I am thinking about adding a 3458A to my list of 'things to think about' and
was wondering if there were any suggestions about what things to consider
when trying to find one.

Any serial number ranges preferred?  Any way to find out when a particular
unit was manufactured?  'HP' or 'Agilent' on the front preferred?  Options
001 or 002 useful or desirable?

I have looked at the Agilent website and the five 'Modification Recommended'
Service Notes.  Also, someone told me that older units could easily add
Option 001 by simply plugging in additional (or different?) memory chips to
available sockets on the board but later units used surface mount chips and
then it was not possible to do that.  Perhaps a return to Agilent or
swapping a board would allow that.

I presume the addition of Option 002 would be simply a change of the
reference board with a re-calibration.

I would appreciate any information about what to consider and where to look
when shopping for a 3458A.

Joe

I am thinking about adding a 3458A to my list of 'things to think about' and was wondering if there were any suggestions about what things to consider when trying to find one. Any serial number ranges preferred? Any way to find out when a particular unit was manufactured? 'HP' or 'Agilent' on the front preferred? Options 001 or 002 useful or desirable? I have looked at the Agilent website and the five 'Modification Recommended' Service Notes. Also, someone told me that older units could easily add Option 001 by simply plugging in additional (or different?) memory chips to available sockets on the board but later units used surface mount chips and then it was not possible to do that. Perhaps a return to Agilent or swapping a board would allow that. I presume the addition of Option 002 would be simply a change of the reference board with a re-calibration. I would appreciate any information about what to consider and where to look when shopping for a 3458A. Joe
PK
Poul-Henning Kamp
Wed, Sep 21, 2011 12:53 PM

In message C876E529248E41D2B39EA89D711B0F8F@S0028384766, "J. L. Trantham" writes:

I am thinking about adding a 3458A to my list of 'things to think about' and
was wondering if there were any suggestions about what things to consider
when trying to find one.

Make sure it is calibrated and that the calibration constants have
not been lost due to old NVRAMS.

Back up CALRAM asap. when you receive unit.

Extra memory is just a couple of standard static RAM chips don't bother.

I presume the addition of Option 002 would be simply a change of the
reference board with a re-calibration.

You probably don't need option 002 since your unit is pre-aged.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <C876E529248E41D2B39EA89D711B0F8F@S0028384766>, "J. L. Trantham" writes: >I am thinking about adding a 3458A to my list of 'things to think about' and >was wondering if there were any suggestions about what things to consider >when trying to find one. Make sure it is calibrated and that the calibration constants have not been lost due to old NVRAMS. Back up CALRAM asap. when you receive unit. Extra memory is just a couple of standard static RAM chips don't bother. >I presume the addition of Option 002 would be simply a change of the >reference board with a re-calibration. You probably don't need option 002 since your unit is pre-aged. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
A
Artekmedia
Wed, Sep 21, 2011 2:06 PM

Joe

The 3458A is still under support by Agilent ...you imply below that you
might consider returning a later unit to them for upgrade ( and or
calibration) . If your seeking to make a significant investment in a
unit because you REALLY NEED all those zero's to the right of the
decimal point I would  contact them and get pricing on the cost of
having the option 002 added and also the cost of a full calibration.
These wont be cheap but at least you will have the $$$$ amount in the
back of your mind so you when you go to purchase a used unit and you can
add these costs to the "total cost of ownership" in making your
decision. While your on the phone be sure and ask them how much longer
they intend to support the 3458A seems like it has been in production
for 15 or 20 years so could be nearing the end of life?

I would also ask this  question to ask over on the Hp_Agilent _Equipment
group on yahoo. I don't recall in recent history seeing any discussion
one way or the other over there about the 3458A which is actually a good
sign as I would infer that the 3458A has very few "warts"

Dave
NR1DX

On 9/21/2011 7:49 AM, J. L. Trantham wrote:

I am thinking about adding a 3458A to my list of 'things to think about' and
was wondering if there were any suggestions about what things to consider
when trying to find one.

Any serial number ranges preferred?  Any way to find out when a particular
unit was manufactured?  'HP' or 'Agilent' on the front preferred?  Options
001 or 002 useful or desirable?

I have looked at the Agilent website and the five 'Modification Recommended'
Service Notes.  Also, someone told me that older units could easily add
Option 001 by simply plugging in additional (or different?) memory chips to
available sockets on the board but later units used surface mount chips and
then it was not possible to do that.  Perhaps a return to Agilent or
swapping a board would allow that.

I presume the addition of Option 002 would be simply a change of the
reference board with a re-calibration.

I would appreciate any information about what to consider and where to look
when shopping for a 3458A.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Dave&  Lynn Henderson
Manuals@ArtekMedia.com
www.Artekmedia.com
PO Box 175
Welch,MN 55089

Joe The 3458A is still under support by Agilent ...you imply below that you might consider returning a later unit to them for upgrade ( and or calibration) . If your seeking to make a significant investment in a unit because you REALLY NEED all those zero's to the right of the decimal point I would contact them and get pricing on the cost of having the option 002 added and also the cost of a full calibration. These wont be cheap but at least you will have the $$$$ amount in the back of your mind so you when you go to purchase a used unit and you can add these costs to the "total cost of ownership" in making your decision. While your on the phone be sure and ask them how much longer they intend to support the 3458A seems like it has been in production for 15 or 20 years so could be nearing the end of life? I would also ask this question to ask over on the Hp_Agilent _Equipment group on yahoo. I don't recall in recent history seeing any discussion one way or the other over there about the 3458A which is actually a good sign as I would infer that the 3458A has very few "warts" Dave NR1DX On 9/21/2011 7:49 AM, J. L. Trantham wrote: > I am thinking about adding a 3458A to my list of 'things to think about' and > was wondering if there were any suggestions about what things to consider > when trying to find one. > > Any serial number ranges preferred? Any way to find out when a particular > unit was manufactured? 'HP' or 'Agilent' on the front preferred? Options > 001 or 002 useful or desirable? > > I have looked at the Agilent website and the five 'Modification Recommended' > Service Notes. Also, someone told me that older units could easily add > Option 001 by simply plugging in additional (or different?) memory chips to > available sockets on the board but later units used surface mount chips and > then it was not possible to do that. Perhaps a return to Agilent or > swapping a board would allow that. > > I presume the addition of Option 002 would be simply a change of the > reference board with a re-calibration. > > I would appreciate any information about what to consider and where to look > when shopping for a 3458A. > > Joe > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > -- Dave& Lynn Henderson Manuals@ArtekMedia.com www.Artekmedia.com PO Box 175 Welch,MN 55089
G
gbusg
Wed, Sep 21, 2011 8:15 PM

Re: Purchase of a used HP 3458A

Here are some things to consider:

  1. Because the 3458A lifetime has already spanned 21 years, date of
    manufacture has a big influence on the real worth of a used 3458A, IMHO. HP
    / Agilent made many revisions to the 3458A over that time, including
    re-design of some boards. In particular the A5 Outboard Controller board was
    re-designed several times.

  2. The s/n prefix indicates date of major revision, but this isn't a
    reliable way to estimate date of manufacture because HP / Agilent sometimes
    went very long times (years) between major revisions. For a better estimate
    of date of manufacturer, look at the date codes on the various I.C.s on
    various boards. Look for the newest date code you can find but excluding
    date codes on a specific board if its date codes are radically different
    than for other boards in the unit, hence the clue that particular board
    might have been replaced at some time. Another way to estimate the date of
    manufacture is to examine the 3458A's serial number suffix. For example
    serial numbers in the low 100s were made in 1988. Serial numbers in the tens
    of thousands were made many years later, after many hardware and firmware
    revisions. It might be nice to look for serial numbers above, say, 8000,
    although for the right price I'd certainly go for a lower serial number. The
    point is that older units are worth less than newer units, IMO.

  3. As Poul-Henning points out, Option 002 probably isn't necessary because
    most non-Option 002 units have aged to the point that they now significantly
    exceed Option 002 specs. (Option 002 7V Reference drift was spec'd 4ppm /
    year, but it's my experience that even non-Option 002 units drift
    significantly less than that. Option 002 is a hand-selected LTZ1000 with low
    drift.)

  4. Look for a unit with Firmware Rev 8 or above. You can determine this via
    the front panel command "Rev?". For the right price a unit having a lower
    Firmware Rev would still be a good deal, but the point is that the low
    Firmware Rev would be another sign of an older unit, probably having some
    older Rev boards and older parts inside as well. Note: In many cases an old
    unit might have been retrofitted with newer firmware, so later Firmware
    Revs, alone, are not proof of the date of manufacture.

  5. The older A5 Controller boards used three Dallas Semiconductor NVRAM to
    store Cal Constants and DMM stored states. Those NVRAM I.C.s have
    non-replaceable internal batteries that were spec'd to last 10 years (but in
    practice probably last something like 14 years). If you end up with an older
    version A5 Controller board, as Poul-Henning points out, retrieve the
    constants in your old NVRAM before their batteries crash and it's too late.
    It's somewhat expensive and inconvenient to replace those three I.C.s, so a
    much newer unit having the latest version A5 board (HP p/n 03458-66547) is
    highly desirable and worth significantly more than older units, IMO. Note:
    older units can be retrofitted with the newer version A5 Controller by
    ordering Agilent Update Kit p/n -3458-80047. Note that the newer version A5
    Controller is completely redesigned and has user replaceable batteries.

  6. Be sure to perform the Self Test from the front panel. The meter must
    pass Self Test or you're probably looking at a very expensive repair.

  7. Additionally, if possible run a "Time Drift" test per Service Note
    3458A-18 to make sure your A3 A/D board is OK. If the meter fails this test,
    you're probably looking at an expensive A3 board repair.

Cheers!
Greg

----- Original Message -----
From: "J. L. Trantham" jltran@att.net
To: "'Discussion of precise voltage measurement'" volt-nuts@febo.com
Sent: Wednesday, September 21, 2011 6:49 AM
Subject: [volt-nuts] 3458A Questions

I am thinking about adding a 3458A to my list of 'things to think about' and
was wondering if there were any suggestions about what things to consider
when trying to find one.

Any serial number ranges preferred?  Any way to find out when a particular
unit was manufactured?  'HP' or 'Agilent' on the front preferred?  Options
001 or 002 useful or desirable?

I have looked at the Agilent website and the five 'Modification Recommended'
Service Notes.  Also, someone told me that older units could easily add
Option 001 by simply plugging in additional (or different?) memory chips to
available sockets on the board but later units used surface mount chips and
then it was not possible to do that.  Perhaps a return to Agilent or
swapping a board would allow that.

I presume the addition of Option 002 would be simply a change of the
reference board with a re-calibration.

I would appreciate any information about what to consider and where to look
when shopping for a 3458A.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: Purchase of a used HP 3458A Here are some things to consider: 1. Because the 3458A lifetime has already spanned 21 years, date of manufacture has a big influence on the real worth of a used 3458A, IMHO. HP / Agilent made many revisions to the 3458A over that time, including re-design of some boards. In particular the A5 Outboard Controller board was re-designed several times. 2. The s/n prefix indicates date of major revision, but this isn't a reliable way to estimate date of manufacture because HP / Agilent sometimes went very long times (years) between major revisions. For a better estimate of date of manufacturer, look at the date codes on the various I.C.s on various boards. Look for the *newest* date code you can find but excluding date codes on a specific board if its date codes are radically different than for other boards in the unit, hence the clue that particular board might have been replaced at some time. Another way to estimate the date of manufacture is to examine the 3458A's serial number *suffix*. For example serial numbers in the low 100s were made in 1988. Serial numbers in the tens of thousands were made many years later, after many hardware and firmware revisions. It might be nice to look for serial numbers above, say, 8000, although for the right price I'd certainly go for a lower serial number. The point is that older units are worth less than newer units, IMO. 3. As Poul-Henning points out, Option 002 probably isn't necessary because most non-Option 002 units have aged to the point that they now significantly exceed Option 002 specs. (Option 002 7V Reference drift was spec'd 4ppm / year, but it's my experience that even non-Option 002 units drift significantly less than that. Option 002 is a hand-selected LTZ1000 with low drift.) 4. Look for a unit with Firmware Rev 8 or above. You can determine this via the front panel command "Rev?". For the right price a unit having a lower Firmware Rev would still be a good deal, but the point is that the low Firmware Rev would be another sign of an older unit, probably having some older Rev boards and older parts inside as well. Note: In many cases an old unit might have been retrofitted with newer firmware, so later Firmware Revs, alone, are not proof of the date of manufacture. 5. The older A5 Controller boards used three Dallas Semiconductor NVRAM to store Cal Constants and DMM stored states. Those NVRAM I.C.s have non-replaceable internal batteries that were spec'd to last 10 years (but in practice probably last something like 14 years). If you end up with an older version A5 Controller board, as Poul-Henning points out, retrieve the constants in your old NVRAM before their batteries crash and it's too late. It's somewhat expensive and inconvenient to replace those three I.C.s, so a much newer unit having the latest version A5 board (HP p/n 03458-66547) is highly desirable and worth significantly *more* than older units, IMO. Note: older units can be retrofitted with the newer version A5 Controller by ordering Agilent Update Kit p/n -3458-80047. Note that the newer version A5 Controller is completely redesigned and has user replaceable batteries. 6. Be sure to perform the Self Test from the front panel. The meter *must* pass Self Test or you're probably looking at a very expensive repair. 7. Additionally, if possible run a "Time Drift" test per Service Note 3458A-18 to make sure your A3 A/D board is OK. If the meter fails this test, you're probably looking at an expensive A3 board repair. Cheers! Greg ----- Original Message ----- From: "J. L. Trantham" <jltran@att.net> To: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com> Sent: Wednesday, September 21, 2011 6:49 AM Subject: [volt-nuts] 3458A Questions I am thinking about adding a 3458A to my list of 'things to think about' and was wondering if there were any suggestions about what things to consider when trying to find one. Any serial number ranges preferred? Any way to find out when a particular unit was manufactured? 'HP' or 'Agilent' on the front preferred? Options 001 or 002 useful or desirable? I have looked at the Agilent website and the five 'Modification Recommended' Service Notes. Also, someone told me that older units could easily add Option 001 by simply plugging in additional (or different?) memory chips to available sockets on the board but later units used surface mount chips and then it was not possible to do that. Perhaps a return to Agilent or swapping a board would allow that. I presume the addition of Option 002 would be simply a change of the reference board with a re-calibration. I would appreciate any information about what to consider and where to look when shopping for a 3458A. Joe _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
PK
Poul-Henning Kamp
Wed, Sep 21, 2011 8:31 PM

In message 00ca01cc789b$2da517c0$6501a8c0@gb02, "gbusg" writes:

Note: In many cases an old
unit might have been retrofitted with newer firmware, so later Firmware
Revs, alone, are not proof of the date of manufacture.

Updating firmware up to and including 8 is just a matter of flipping
eproms, something you can easily do yourself.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <00ca01cc789b$2da517c0$6501a8c0@gb02>, "gbusg" writes: >Note: In many cases an old >unit might have been retrofitted with newer firmware, so later Firmware >Revs, alone, are not proof of the date of manufacture. Updating firmware up to and including 8 is just a matter of flipping eproms, something you can easily do yourself. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.