Jerry Hancock
Mon, Jan 4, 2016 2:26 AM
I’ve been reading the list for a while, never had a reason to post until now.
I have an HP 3457a, actually two of them, both work great, accurate, stable, etc., except one won’t autocal without an error 512 in auxerr - ohms recharge offset. I’ve had the meters side by side for hours the past few days and can’t seem to locate the problem. There is a difference in the Ohms precharge offset in the failing meter. It looks like a factor mod was performed to lower U216 below +15 by adding resistors to legs 13 and 10. I know there were service notes sent out by HP in 1992 related to the 512 error. the other thing of note is that occasionally when I power on the meter, it will also get a 16 error, which is the offset. I ran diagnostic 4 and calibrated the offset but occasionally the meter will not pass the test no mater how the offset it adjusted. For those of you not familiar, the diagnostic 4 puts the meter in a loop displaying HI or LO depending on the adjustment until the repair tech (me in this case) adjusts the offset correctly. At that point the meter displays ‘Passed’. When the meter fails, usually when cold, the display oscillates between HI/LO like the ADC doesn’t have a connection and the input is floating. I tried swapping U101, the input hybrid, between the two units without luck.
So I am wondering if anyone on the list has run into this problem in the past, and if so, the solution would be appreciated.
Thanks.
By the way, I have a number of voltmeters, can’t seem to get enough, and am pretty good with 3455A repairs having four of them. My black face 3455A is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x 3478a, 2 x 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and others, mostly older HP units like the 400. I am currently eyeing a 3458a. I guess I am a volt-nut at heart.
I’ve been reading the list for a while, never had a reason to post until now.
I have an HP 3457a, actually two of them, both work great, accurate, stable, etc., except one won’t autocal without an error 512 in auxerr - ohms recharge offset. I’ve had the meters side by side for hours the past few days and can’t seem to locate the problem. There is a difference in the Ohms precharge offset in the failing meter. It looks like a factor mod was performed to lower U216 below +15 by adding resistors to legs 13 and 10. I know there were service notes sent out by HP in 1992 related to the 512 error. the other thing of note is that occasionally when I power on the meter, it will also get a 16 error, which is the offset. I ran diagnostic 4 and calibrated the offset but occasionally the meter will not pass the test no mater how the offset it adjusted. For those of you not familiar, the diagnostic 4 puts the meter in a loop displaying HI or LO depending on the adjustment until the repair tech (me in this case) adjusts the offset correctly. At that point the meter displays ‘Passed’. When the meter fails, usually when cold, the display oscillates between HI/LO like the ADC doesn’t have a connection and the input is floating. I tried swapping U101, the input hybrid, between the two units without luck.
So I am wondering if anyone on the list has run into this problem in the past, and if so, the solution would be appreciated.
Thanks.
By the way, I have a number of voltmeters, can’t seem to get enough, and am pretty good with 3455A repairs having four of them. My black face 3455A is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x 3478a, 2 x 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and others, mostly older HP units like the 400. I am currently eyeing a 3458a. I guess I am a volt-nut at heart.
JH
Jerry Hancock
Wed, Jan 6, 2016 1:06 AM
Hello again, Just for the record, after spending another dozen or so hours on this problem, I found U216, “Precharge offset adj” to have a low resistance from pins 1,5,6 to pin 4, essentially one of the 4 transistors in the CA3096 is shorted (11.28 ohms) though not completely. I found this by scoping the signal coming into the input amplifier at pin 10 of U101, “Input Hybrid”, while putting the meter into an auto cal loop. This problem was difficult to track down as unlike the other Service Manuals, there are no waveforms available in the 3457a manual. I noticed that during the precharge cycle the waveform was taking a large negative dive instead of two relatively equal pulses, one that reads the signal and the other during precharge, I assume. This negative voltage from the precharge offset was then not being drained efficiently when the leads were open resulting in a negative to zero open voltage displayed on the meter compared to the typical ~+2.5V.
So for the record, if you get a “0 Hardware error” followed by a 512 or 560 (512 + 32 + 16 errors signaled) in reply to the “auxerr?” command, take a look in the area of U216 and check the four transistors that are in the CA3096E.
Lastly, as I commented on another site, after reading through the 3457a theory of operations, that there are a number factors that need to be calibrated out of a meter to ensure accuracy and repeatable measurements. In this case, obviously the input amplifier needs to have a stable offset voltage (zero). The precharge offset charges the approximate 10pf of stray capacitance out of the input amplifier to keep it from impacting the readings. In conclusion, it takes more than an accurate ADC, gain resistors, etc to make a good meter. I followed a number of threads where people were attempting to make their own 5.5 digit meter, I played around with some circuits myself, but nowhere did I see stray capacitance taken into account.
Thank you for reading. By the way, there is what looks like a nice 3458a on craigslist in the SF Bay area. Initial price of $2,900 was dropped last night to $1,950. I wish I had some leftover cash from Christmas!
Jerry
On Jan 3, 2016, at 6:26 PM, Jerry Hancock jerry@hanler.com wrote:
I’ve been reading the list for a while, never had a reason to post until now.
I have an HP 3457a, actually two of them, both work great, accurate, stable, etc., except one won’t autocal without an error 512 in auxerr - ohms recharge offset. I’ve had the meters side by side for hours the past few days and can’t seem to locate the problem. There is a difference in the Ohms precharge offset in the failing meter. It looks like a factor mod was performed to lower U216 below +15 by adding resistors to legs 13 and 10. I know there were service notes sent out by HP in 1992 related to the 512 error. the other thing of note is that occasionally when I power on the meter, it will also get a 16 error, which is the offset. I ran diagnostic 4 and calibrated the offset but occasionally the meter will not pass the test no mater how the offset it adjusted. For those of you not familiar, the diagnostic 4 puts the meter in a loop displaying HI or LO depending on the adjustment until the repair tech (me in this case) adjusts the offset correctly. At that point the meter displays ‘Passed’. When the meter fails, usually when cold, the display oscillates between HI/LO like the ADC doesn’t have a connection and the input is floating. I tried swapping U101, the input hybrid, between the two units without luck.
So I am wondering if anyone on the list has run into this problem in the past, and if so, the solution would be appreciated.
Thanks.
By the way, I have a number of voltmeters, can’t seem to get enough, and am pretty good with 3455A repairs having four of them. My black face 3455A is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x 3478a, 2 x 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and others, mostly older HP units like the 400. I am currently eyeing a 3458a. I guess I am a volt-nut at heart.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello again, Just for the record, after spending another dozen or so hours on this problem, I found U216, “Precharge offset adj” to have a low resistance from pins 1,5,6 to pin 4, essentially one of the 4 transistors in the CA3096 is shorted (11.28 ohms) though not completely. I found this by scoping the signal coming into the input amplifier at pin 10 of U101, “Input Hybrid”, while putting the meter into an auto cal loop. This problem was difficult to track down as unlike the other Service Manuals, there are no waveforms available in the 3457a manual. I noticed that during the precharge cycle the waveform was taking a large negative dive instead of two relatively equal pulses, one that reads the signal and the other during precharge, I assume. This negative voltage from the precharge offset was then not being drained efficiently when the leads were open resulting in a negative to zero open voltage displayed on the meter compared to the typical ~+2.5V.
So for the record, if you get a “0 Hardware error” followed by a 512 or 560 (512 + 32 + 16 errors signaled) in reply to the “auxerr?” command, take a look in the area of U216 and check the four transistors that are in the CA3096E.
Lastly, as I commented on another site, after reading through the 3457a theory of operations, that there are a number factors that need to be calibrated out of a meter to ensure accuracy and repeatable measurements. In this case, obviously the input amplifier needs to have a stable offset voltage (zero). The precharge offset charges the approximate 10pf of stray capacitance out of the input amplifier to keep it from impacting the readings. In conclusion, it takes more than an accurate ADC, gain resistors, etc to make a good meter. I followed a number of threads where people were attempting to make their own 5.5 digit meter, I played around with some circuits myself, but nowhere did I see stray capacitance taken into account.
Thank you for reading. By the way, there is what looks like a nice 3458a on craigslist in the SF Bay area. Initial price of $2,900 was dropped last night to $1,950. I wish I had some leftover cash from Christmas!
Jerry
On Jan 3, 2016, at 6:26 PM, Jerry Hancock <jerry@hanler.com> wrote:
> I’ve been reading the list for a while, never had a reason to post until now.
>
> I have an HP 3457a, actually two of them, both work great, accurate, stable, etc., except one won’t autocal without an error 512 in auxerr - ohms recharge offset. I’ve had the meters side by side for hours the past few days and can’t seem to locate the problem. There is a difference in the Ohms precharge offset in the failing meter. It looks like a factor mod was performed to lower U216 below +15 by adding resistors to legs 13 and 10. I know there were service notes sent out by HP in 1992 related to the 512 error. the other thing of note is that occasionally when I power on the meter, it will also get a 16 error, which is the offset. I ran diagnostic 4 and calibrated the offset but occasionally the meter will not pass the test no mater how the offset it adjusted. For those of you not familiar, the diagnostic 4 puts the meter in a loop displaying HI or LO depending on the adjustment until the repair tech (me in this case) adjusts the offset correctly. At that point the meter displays ‘Passed’. When the meter fails, usually when cold, the display oscillates between HI/LO like the ADC doesn’t have a connection and the input is floating. I tried swapping U101, the input hybrid, between the two units without luck.
>
> So I am wondering if anyone on the list has run into this problem in the past, and if so, the solution would be appreciated.
>
> Thanks.
>
> By the way, I have a number of voltmeters, can’t seem to get enough, and am pretty good with 3455A repairs having four of them. My black face 3455A is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x 3478a, 2 x 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and others, mostly older HP units like the 400. I am currently eyeing a 3458a. I guess I am a volt-nut at heart.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
TM
Tom Miller
Wed, Jan 6, 2016 1:28 AM
That looks like a good one from what I can see. It would be nice to check it
out in person to see if the display has any dim digits. Looking at the
picture seems to show it has some dimming compared to the manual range
indicator. The price is pretty good also. Maybe they might negotiate down
slightly.
It would run you about $600+ to send in for a Keysight calibration.
Nice work on finding the problem with your 3457A.
----- Original Message -----
From: "Jerry Hancock" jerry@hanler.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Tuesday, January 05, 2016 8:06 PM
Subject: Re: [volt-nuts] New member with 3457A problem
Hello again, Just for the record, after spending another dozen or so hours
on this problem, I found U216, “Precharge offset adj” to have a low
resistance from pins 1,5,6 to pin 4, essentially one of the 4 transistors
in the CA3096 is shorted (11.28 ohms) though not completely. I found this
by scoping the signal coming into the input amplifier at pin 10 of U101,
“Input Hybrid”, while putting the meter into an auto cal loop. This
problem was difficult to track down as unlike the other Service Manuals,
there are no waveforms available in the 3457a manual. I noticed that
during the precharge cycle the waveform was taking a large negative dive
instead of two relatively equal pulses, one that reads the signal and the
other during precharge, I assume. This negative voltage from the
precharge offset was then not being drained efficiently when the leads
were open resulting in a negative to zero open voltage displayed on the
meter compared to the typical ~+2.5V.
So for the record, if you get a “0 Hardware error” followed by a 512 or
560 (512 + 32 + 16 errors signaled) in reply to the “auxerr?” command,
take a look in the area of U216 and check the four transistors that are in
the CA3096E.
Lastly, as I commented on another site, after reading through the 3457a
theory of operations, that there are a number factors that need to be
calibrated out of a meter to ensure accuracy and repeatable measurements.
In this case, obviously the input amplifier needs to have a stable offset
voltage (zero). The precharge offset charges the approximate 10pf of
stray capacitance out of the input amplifier to keep it from impacting the
readings. In conclusion, it takes more than an accurate ADC, gain
resistors, etc to make a good meter. I followed a number of threads where
people were attempting to make their own 5.5 digit meter, I played around
with some circuits myself, but nowhere did I see stray capacitance taken
into account.
Thank you for reading. By the way, there is what looks like a nice 3458a
on craigslist in the SF Bay area. Initial price of $2,900 was dropped
last night to $1,950. I wish I had some leftover cash from Christmas!
Jerry
On Jan 3, 2016, at 6:26 PM, Jerry Hancock jerry@hanler.com wrote:
I’ve been reading the list for a while, never had a reason to post until
now.
I have an HP 3457a, actually two of them, both work great, accurate,
stable, etc., except one won’t autocal without an error 512 in auxerr -
ohms recharge offset. I’ve had the meters side by side for hours the
past few days and can’t seem to locate the problem. There is a
difference in the Ohms precharge offset in the failing meter. It looks
like a factor mod was performed to lower U216 below +15 by adding
resistors to legs 13 and 10. I know there were service notes sent out by
HP in 1992 related to the 512 error. the other thing of note is that
occasionally when I power on the meter, it will also get a 16 error,
which is the offset. I ran diagnostic 4 and calibrated the offset but
occasionally the meter will not pass the test no mater how the offset it
adjusted. For those of you not familiar, the diagnostic 4 puts the meter
in a loop displaying HI or LO depending on the adjustment until the
repair tech (me in this case) adjusts the offset correctly. At that
point the
meter displays ‘Passed’. When the meter fails, usually when cold, the
display oscillates between HI/LO like the ADC doesn’t have a connection
and the input is floating. I tried swapping U101, the input hybrid,
between the two units without luck.
So I am wondering if anyone on the list has run into this problem in the
past, and if so, the solution would be appreciated.
Thanks.
By the way, I have a number of voltmeters, can’t seem to get enough, and
am pretty good with 3455A repairs having four of them. My black face
3455A is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x
3478a, 2 x 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and
others, mostly older HP units like the 400. I am currently eyeing a
3458a. I guess I am a volt-nut at heart.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
That looks like a good one from what I can see. It would be nice to check it
out in person to see if the display has any dim digits. Looking at the
picture seems to show it has some dimming compared to the manual range
indicator. The price is pretty good also. Maybe they might negotiate down
slightly.
It would run you about $600+ to send in for a Keysight calibration.
Nice work on finding the problem with your 3457A.
----- Original Message -----
From: "Jerry Hancock" <jerry@hanler.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Tuesday, January 05, 2016 8:06 PM
Subject: Re: [volt-nuts] New member with 3457A problem
> Hello again, Just for the record, after spending another dozen or so hours
> on this problem, I found U216, “Precharge offset adj” to have a low
> resistance from pins 1,5,6 to pin 4, essentially one of the 4 transistors
> in the CA3096 is shorted (11.28 ohms) though not completely. I found this
> by scoping the signal coming into the input amplifier at pin 10 of U101,
> “Input Hybrid”, while putting the meter into an auto cal loop. This
> problem was difficult to track down as unlike the other Service Manuals,
> there are no waveforms available in the 3457a manual. I noticed that
> during the precharge cycle the waveform was taking a large negative dive
> instead of two relatively equal pulses, one that reads the signal and the
> other during precharge, I assume. This negative voltage from the
> precharge offset was then not being drained efficiently when the leads
> were open resulting in a negative to zero open voltage displayed on the
> meter compared to the typical ~+2.5V.
>
> So for the record, if you get a “0 Hardware error” followed by a 512 or
> 560 (512 + 32 + 16 errors signaled) in reply to the “auxerr?” command,
> take a look in the area of U216 and check the four transistors that are in
> the CA3096E.
>
> Lastly, as I commented on another site, after reading through the 3457a
> theory of operations, that there are a number factors that need to be
> calibrated out of a meter to ensure accuracy and repeatable measurements.
> In this case, obviously the input amplifier needs to have a stable offset
> voltage (zero). The precharge offset charges the approximate 10pf of
> stray capacitance out of the input amplifier to keep it from impacting the
> readings. In conclusion, it takes more than an accurate ADC, gain
> resistors, etc to make a good meter. I followed a number of threads where
> people were attempting to make their own 5.5 digit meter, I played around
> with some circuits myself, but nowhere did I see stray capacitance taken
> into account.
>
> Thank you for reading. By the way, there is what looks like a nice 3458a
> on craigslist in the SF Bay area. Initial price of $2,900 was dropped
> last night to $1,950. I wish I had some leftover cash from Christmas!
>
> Jerry
>
> On Jan 3, 2016, at 6:26 PM, Jerry Hancock <jerry@hanler.com> wrote:
>
>> I’ve been reading the list for a while, never had a reason to post until
>> now.
>>
>> I have an HP 3457a, actually two of them, both work great, accurate,
>> stable, etc., except one won’t autocal without an error 512 in auxerr -
>> ohms recharge offset. I’ve had the meters side by side for hours the
>> past few days and can’t seem to locate the problem. There is a
>> difference in the Ohms precharge offset in the failing meter. It looks
>> like a factor mod was performed to lower U216 below +15 by adding
>> resistors to legs 13 and 10. I know there were service notes sent out by
>> HP in 1992 related to the 512 error. the other thing of note is that
>> occasionally when I power on the meter, it will also get a 16 error,
>> which is the offset. I ran diagnostic 4 and calibrated the offset but
>> occasionally the meter will not pass the test no mater how the offset it
>> adjusted. For those of you not familiar, the diagnostic 4 puts the meter
>> in a loop displaying HI or LO depending on the adjustment until the
>> repair tech (me in this case) adjusts the offset correctly. At that
>> point the
> meter displays ‘Passed’. When the meter fails, usually when cold, the
> display oscillates between HI/LO like the ADC doesn’t have a connection
> and the input is floating. I tried swapping U101, the input hybrid,
> between the two units without luck.
>>
>> So I am wondering if anyone on the list has run into this problem in the
>> past, and if so, the solution would be appreciated.
>>
>> Thanks.
>>
>> By the way, I have a number of voltmeters, can’t seem to get enough, and
>> am pretty good with 3455A repairs having four of them. My black face
>> 3455A is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x
>> 3478a, 2 x 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and
>> others, mostly older HP units like the 400. I am currently eyeing a
>> 3458a. I guess I am a volt-nut at heart.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
PL
Pete Lancashire
Wed, Jan 6, 2016 3:42 AM
Seen a lot worse even on rentals. Pretty much the same here, other big $
projects going on right now so can't make an offer.
On Tue, Jan 5, 2016 at 5:28 PM, Tom Miller tmiller11147@verizon.net wrote:
That looks like a good one from what I can see. It would be nice to check
it out in person to see if the display has any dim digits. Looking at the
picture seems to show it has some dimming compared to the manual range
indicator. The price is pretty good also. Maybe they might negotiate down
slightly.
It would run you about $600+ to send in for a Keysight calibration.
Nice work on finding the problem with your 3457A.
----- Original Message ----- From: "Jerry Hancock" jerry@hanler.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Tuesday, January 05, 2016 8:06 PM
Subject: Re: [volt-nuts] New member with 3457A problem
Hello again, Just for the record, after spending another dozen or so hours
on this problem, I found U216, “Precharge offset adj” to have a low
resistance from pins 1,5,6 to pin 4, essentially one of the 4 transistors
in the CA3096 is shorted (11.28 ohms) though not completely. I found this
by scoping the signal coming into the input amplifier at pin 10 of U101,
“Input Hybrid”, while putting the meter into an auto cal loop. This
problem was difficult to track down as unlike the other Service Manuals,
there are no waveforms available in the 3457a manual. I noticed that
during the precharge cycle the waveform was taking a large negative dive
instead of two relatively equal pulses, one that reads the signal and the
other during precharge, I assume. This negative voltage from the precharge
offset was then not being drained efficiently when the leads were open
resulting in a negative to zero open voltage displayed on the meter
compared to the typical ~+2.5V.
So for the record, if you get a “0 Hardware error” followed by a 512 or
560 (512 + 32 + 16 errors signaled) in reply to the “auxerr?” command, take
a look in the area of U216 and check the four transistors that are in the
CA3096E.
Lastly, as I commented on another site, after reading through the 3457a
theory of operations, that there are a number factors that need to be
calibrated out of a meter to ensure accuracy and repeatable measurements.
In this case, obviously the input amplifier needs to have a stable offset
voltage (zero). The precharge offset charges the approximate 10pf of stray
capacitance out of the input amplifier to keep it from impacting the
readings. In conclusion, it takes more than an accurate ADC, gain
resistors, etc to make a good meter. I followed a number of threads where
people were attempting to make their own 5.5 digit meter, I played around
with some circuits myself, but nowhere did I see stray capacitance taken
into account.
Thank you for reading. By the way, there is what looks like a nice 3458a
on craigslist in the SF Bay area. Initial price of $2,900 was dropped last
night to $1,950. I wish I had some leftover cash from Christmas!
Jerry
On Jan 3, 2016, at 6:26 PM, Jerry Hancock jerry@hanler.com wrote:
I’ve been reading the list for a while, never had a reason to post until
now.
I have an HP 3457a, actually two of them, both work great, accurate,
stable, etc., except one won’t autocal without an error 512 in auxerr -
ohms recharge offset. I’ve had the meters side by side for hours the past
few days and can’t seem to locate the problem. There is a difference in
the Ohms precharge offset in the failing meter. It looks like a factor mod
was performed to lower U216 below +15 by adding resistors to legs 13 and
10. I know there were service notes sent out by HP in 1992 related to the
512 error. the other thing of note is that occasionally when I power on the
meter, it will also get a 16 error, which is the offset. I ran diagnostic
4 and calibrated the offset but occasionally the meter will not pass the
test no mater how the offset it adjusted. For those of you not familiar,
the diagnostic 4 puts the meter in a loop displaying HI or LO depending on
the adjustment until the repair tech (me in this case) adjusts the offset
correctly. At that point the
meter displays ‘Passed’. When the meter fails, usually when cold, the
display oscillates between HI/LO like the ADC doesn’t have a connection and
the input is floating. I tried swapping U101, the input hybrid, between
the two units without luck.
So I am wondering if anyone on the list has run into this problem in the
past, and if so, the solution would be appreciated.
Thanks.
By the way, I have a number of voltmeters, can’t seem to get enough, and
am pretty good with 3455A repairs having four of them. My black face 3455A
is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x 3478a, 2 x
3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and others, mostly
older HP units like the 400. I am currently eyeing a 3458a. I guess I am
a volt-nut at heart.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Seen a lot worse even on rentals. Pretty much the same here, other big $
projects going on right now so can't make an offer.
On Tue, Jan 5, 2016 at 5:28 PM, Tom Miller <tmiller11147@verizon.net> wrote:
> That looks like a good one from what I can see. It would be nice to check
> it out in person to see if the display has any dim digits. Looking at the
> picture seems to show it has some dimming compared to the manual range
> indicator. The price is pretty good also. Maybe they might negotiate down
> slightly.
>
> It would run you about $600+ to send in for a Keysight calibration.
>
> Nice work on finding the problem with your 3457A.
>
> ----- Original Message ----- From: "Jerry Hancock" <jerry@hanler.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Tuesday, January 05, 2016 8:06 PM
> Subject: Re: [volt-nuts] New member with 3457A problem
>
>
>
> Hello again, Just for the record, after spending another dozen or so hours
>> on this problem, I found U216, “Precharge offset adj” to have a low
>> resistance from pins 1,5,6 to pin 4, essentially one of the 4 transistors
>> in the CA3096 is shorted (11.28 ohms) though not completely. I found this
>> by scoping the signal coming into the input amplifier at pin 10 of U101,
>> “Input Hybrid”, while putting the meter into an auto cal loop. This
>> problem was difficult to track down as unlike the other Service Manuals,
>> there are no waveforms available in the 3457a manual. I noticed that
>> during the precharge cycle the waveform was taking a large negative dive
>> instead of two relatively equal pulses, one that reads the signal and the
>> other during precharge, I assume. This negative voltage from the precharge
>> offset was then not being drained efficiently when the leads were open
>> resulting in a negative to zero open voltage displayed on the meter
>> compared to the typical ~+2.5V.
>>
>> So for the record, if you get a “0 Hardware error” followed by a 512 or
>> 560 (512 + 32 + 16 errors signaled) in reply to the “auxerr?” command, take
>> a look in the area of U216 and check the four transistors that are in the
>> CA3096E.
>>
>> Lastly, as I commented on another site, after reading through the 3457a
>> theory of operations, that there are a number factors that need to be
>> calibrated out of a meter to ensure accuracy and repeatable measurements.
>> In this case, obviously the input amplifier needs to have a stable offset
>> voltage (zero). The precharge offset charges the approximate 10pf of stray
>> capacitance out of the input amplifier to keep it from impacting the
>> readings. In conclusion, it takes more than an accurate ADC, gain
>> resistors, etc to make a good meter. I followed a number of threads where
>> people were attempting to make their own 5.5 digit meter, I played around
>> with some circuits myself, but nowhere did I see stray capacitance taken
>> into account.
>>
>> Thank you for reading. By the way, there is what looks like a nice 3458a
>> on craigslist in the SF Bay area. Initial price of $2,900 was dropped last
>> night to $1,950. I wish I had some leftover cash from Christmas!
>>
>> Jerry
>>
>> On Jan 3, 2016, at 6:26 PM, Jerry Hancock <jerry@hanler.com> wrote:
>>
>> I’ve been reading the list for a while, never had a reason to post until
>>> now.
>>>
>>> I have an HP 3457a, actually two of them, both work great, accurate,
>>> stable, etc., except one won’t autocal without an error 512 in auxerr -
>>> ohms recharge offset. I’ve had the meters side by side for hours the past
>>> few days and can’t seem to locate the problem. There is a difference in
>>> the Ohms precharge offset in the failing meter. It looks like a factor mod
>>> was performed to lower U216 below +15 by adding resistors to legs 13 and
>>> 10. I know there were service notes sent out by HP in 1992 related to the
>>> 512 error. the other thing of note is that occasionally when I power on the
>>> meter, it will also get a 16 error, which is the offset. I ran diagnostic
>>> 4 and calibrated the offset but occasionally the meter will not pass the
>>> test no mater how the offset it adjusted. For those of you not familiar,
>>> the diagnostic 4 puts the meter in a loop displaying HI or LO depending on
>>> the adjustment until the repair tech (me in this case) adjusts the offset
>>> correctly. At that point the
>>>
>> meter displays ‘Passed’. When the meter fails, usually when cold, the
>> display oscillates between HI/LO like the ADC doesn’t have a connection and
>> the input is floating. I tried swapping U101, the input hybrid, between
>> the two units without luck.
>>
>>>
>>> So I am wondering if anyone on the list has run into this problem in the
>>> past, and if so, the solution would be appreciated.
>>>
>>> Thanks.
>>>
>>> By the way, I have a number of voltmeters, can’t seem to get enough, and
>>> am pretty good with 3455A repairs having four of them. My black face 3455A
>>> is one of my favorites. I also have 2 x 3456A, 2 x 3457A, 3 x 3478a, 2 x
>>> 3476, a bunch of Keithely (191, 178, 199, 195, 196, etc) and others, mostly
>>> older HP units like the 400. I am currently eyeing a 3458a. I guess I am
>>> a volt-nut at heart.
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>>>
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>