Think twice about those Vishay metal film resistors...  they have known stability problems (particularly concerning humidity).  Also they can have problems meeting their own specs.  There is an active thread of eevblog.com concerning the issue.  Consensus seems to be precision wire wound resistors are a better choice.


These will be replaced by 2 vishay 0.01% resistors which are hopefully more stable.

Think twice about those Vishay metal film resistors... they have known stability problems (particularly concerning humidity). Also they can have problems meeting their own specs. There is an active thread of eevblog.com concerning the issue. Consensus seems to be precision wire wound resistors are a better choice. -------------------- These will be replaced by 2 vishay 0.01% resistors which are hopefully more stable.
TK
Tom Knox
Sat, May 16, 2015 6:56 PM

What is the group consensus on Vishay metal foil?
Thanks

Thomas Knox

From: holrum@hotmail.com
To: volt-nuts@febo.com
Date: Sat, 16 May 2015 18:22:22 +0000
Subject: [volt-nuts] How accurate is a fluke after 9 years

Think twice about those Vishay metal film resistors...  they have known stability problems (particularly concerning humidity).  Also they can have problems meeting their own specs.  There is an active thread of eevblog.com concerning the issue.  Consensus seems to be precision wire wound resistors are a better choice.


These will be replaced by 2 vishay 0.01% resistors which are hopefully more stable.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

What is the group consensus on Vishay metal foil? Thanks Thomas Knox > From: holrum@hotmail.com > To: volt-nuts@febo.com > Date: Sat, 16 May 2015 18:22:22 +0000 > Subject: [volt-nuts] How accurate is a fluke after 9 years > > Think twice about those Vishay metal film resistors... they have known stability problems (particularly concerning humidity). Also they can have problems meeting their own specs. There is an active thread of eevblog.com concerning the issue. Consensus seems to be precision wire wound resistors are a better choice. > > -------------------- > These will be replaced by 2 vishay 0.01% resistors which are hopefully more stable. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
KP
Ken Peek
Sat, May 16, 2015 7:11 PM

IMHO:  Vishay metal-foil resistors are very good, but ONLY if you get
them in a hermetic package.  Depending on where you are in the world,
this can be rather expensive [more or less].  Precision wire-wound
resistors made with with low TCR Evanohm wire THAT ARE PROPERLY WELDED
are the next best thing, and they don't cost you your first born...

IMHO: Vishay metal-foil resistors are very good, but *ONLY* if you get them in a hermetic package. Depending on where you are in the world, this can be rather expensive [more or less]. Precision wire-wound resistors made with with low TCR Evanohm wire *THAT ARE PROPERLY WELDED* are the next best thing, and they don't cost you your first born...
AM
Anton Moehammad
Sun, May 17, 2015 1:40 AM

Hi All,What about Edward G Pettis resistor are they include in THAT ARE PROPERLY WELDED

 Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis:

IMHO:  Vishay metal-foil resistors are very good, but ONLY if you get
them in a hermetic package.  Depending on where you are in the world,
this can be rather expensive [more or less].  Precision wire-wound
resistors made with with low TCR Evanohm wire THAT ARE PROPERLY WELDED
are the next best thing, and they don't cost you your first born...


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi All,What about Edward G Pettis resistor are they include in *THAT ARE PROPERLY WELDED* Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis: IMHO:  Vishay metal-foil resistors are very good, but *ONLY* if you get them in a hermetic package.  Depending on where you are in the world, this can be rather expensive [more or less].  Precision wire-wound resistors made with with low TCR Evanohm wire *THAT ARE PROPERLY WELDED* are the next best thing, and they don't cost you your first born... _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
A
acbern@gmx.de
Sun, May 17, 2015 4:16 PM

I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper.
you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa.
I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5).

Adrian

Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr
Von: "Anton Moehammad via volt-nuts" volt-nuts@febo.com
An: "Ken.Peek@DiligentMinds.com" Ken.Peek@DiligentMinds.com, "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Bls:  How accurate is a fluke after 9 years

Hi All,What about Edward G Pettis resistor are they include in THAT ARE PROPERLY WELDED

  Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis:

IMHO:  Vishay metal-foil resistors are very good, but ONLY if you get
them in a hermetic package.  Depending on where you are in the world,
this can be rather expensive [more or less].  Precision wire-wound
resistors made with with low TCR Evanohm wire THAT ARE PROPERLY WELDED
are the next best thing, and they don't cost you your first born...


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper. you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa. I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5). Adrian > Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr > Von: "Anton Moehammad via volt-nuts" <volt-nuts@febo.com> > An: "Ken.Peek@DiligentMinds.com" <Ken.Peek@DiligentMinds.com>, "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: [volt-nuts] Bls: How accurate is a fluke after 9 years > > Hi All,What about Edward G Pettis resistor are they include in *THAT ARE PROPERLY WELDED* > > > > Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis: > > > IMHO:  Vishay metal-foil resistors are very good, but *ONLY* if you get > them in a hermetic package.  Depending on where you are in the world, > this can be rather expensive [more or less].  Precision wire-wound > resistors made with with low TCR Evanohm wire *THAT ARE PROPERLY WELDED* > are the next best thing, and they don't cost you your first born... > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
AJ
Andreas Jahn
Sun, May 17, 2015 8:00 PM

Hello Adrian,

which actual type of resistor did you use.
And from where did you order it?
For which quantity is the 40 usd price? (single piece?)

I guess it was one of the hzseries of vishay. right?
Perhaps a Y4725 (VHA512Z?)
http://www.vishaypg.com/docs/63120/hzseries.pdf
Was it including PMO?

With best regards

Andreas

Am 17.05.2015 um 18:16 schrieb acbern@gmx.de:

I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper.
you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa.
I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5).

Adrian

Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr
Von: "Anton Moehammad via volt-nuts" volt-nuts@febo.com
An: "Ken.Peek@DiligentMinds.com" Ken.Peek@DiligentMinds.com, "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Bls:  How accurate is a fluke after 9 years

Hi All,What about Edward G Pettis resistor are they include in THAT ARE PROPERLY WELDED

   Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis:
 

IMHO:  Vishay metal-foil resistors are very good, but ONLY if you get
them in a hermetic package.  Depending on where you are in the world,
this can be rather expensive [more or less].  Precision wire-wound
resistors made with with low TCR Evanohm wire THAT ARE PROPERLY WELDED
are the next best thing, and they don't cost you your first born...


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hello Adrian, which actual type of resistor did you use. And from where did you order it? For which quantity is the 40 usd price? (single piece?) I guess it was one of the hzseries of vishay. right? Perhaps a Y4725 (VHA512Z?) http://www.vishaypg.com/docs/63120/hzseries.pdf Was it including PMO? With best regards Andreas Am 17.05.2015 um 18:16 schrieb acbern@gmx.de: > I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper. > you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa. > I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5). > > Adrian > >> Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr >> Von: "Anton Moehammad via volt-nuts" <volt-nuts@febo.com> >> An: "Ken.Peek@DiligentMinds.com" <Ken.Peek@DiligentMinds.com>, "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> Betreff: [volt-nuts] Bls: How accurate is a fluke after 9 years >> >> Hi All,What about Edward G Pettis resistor are they include in *THAT ARE PROPERLY WELDED* >> >> >> >> Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis: >> >> >> IMHO: Vishay metal-foil resistors are very good, but *ONLY* if you get >> them in a hermetic package. Depending on where you are in the world, >> this can be rather expensive [more or less]. Precision wire-wound >> resistors made with with low TCR Evanohm wire *THAT ARE PROPERLY WELDED* >> are the next best thing, and they don't cost you your first born... >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
A
acbern@gmx.de
Sun, May 17, 2015 8:25 PM

hi andreas,

yes, it was a vha512 (maybe 516, but dont think so, dont have data handy). I ordered it from their berlin area based distributor (powertron). I ordered a couple, but different values, so it was single piece price. was some time ago, so price may have increased by now.

adrian

Gesendet: Sonntag, 17. Mai 2015 um 22:00 Uhr
Von: "Andreas Jahn" Andreas_-_Jahn@t-online.de
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Bls:  How accurate is a fluke after 9 years

Hello Adrian,

which actual type of resistor did you use.
And from where did you order it?
For which quantity is the 40 usd price? (single piece?)

I guess it was one of the hzseries of vishay. right?
Perhaps a Y4725 (VHA512Z?)
http://www.vishaypg.com/docs/63120/hzseries.pdf
Was it including PMO?

With best regards

Andreas

Am 17.05.2015 um 18:16 schrieb acbern@gmx.de:

I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper.
you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa.
I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5).

Adrian

Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr
Von: "Anton Moehammad via volt-nuts" volt-nuts@febo.com
An: "Ken.Peek@DiligentMinds.com" Ken.Peek@DiligentMinds.com, "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Bls:  How accurate is a fluke after 9 years

Hi All,What about Edward G Pettis resistor are they include in THAT ARE PROPERLY WELDED

   Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis:
 

IMHO:  Vishay metal-foil resistors are very good, but ONLY if you get
them in a hermetic package.  Depending on where you are in the world,
this can be rather expensive [more or less].  Precision wire-wound
resistors made with with low TCR Evanohm wire THAT ARE PROPERLY WELDED
are the next best thing, and they don't cost you your first born...


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

hi andreas, yes, it was a vha512 (maybe 516, but dont think so, dont have data handy). I ordered it from their berlin area based distributor (powertron). I ordered a couple, but different values, so it was single piece price. was some time ago, so price may have increased by now. adrian > Gesendet: Sonntag, 17. Mai 2015 um 22:00 Uhr > Von: "Andreas Jahn" <Andreas_-_Jahn@t-online.de> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] Bls: How accurate is a fluke after 9 years > > Hello Adrian, > > which actual type of resistor did you use. > And from where did you order it? > For which quantity is the 40 usd price? (single piece?) > > I guess it was one of the hzseries of vishay. right? > Perhaps a Y4725 (VHA512Z?) > http://www.vishaypg.com/docs/63120/hzseries.pdf > Was it including PMO? > > With best regards > > Andreas > > Am 17.05.2015 um 18:16 schrieb acbern@gmx.de: > > I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper. > > you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa. > > I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5). > > > > Adrian > > > >> Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr > >> Von: "Anton Moehammad via volt-nuts" <volt-nuts@febo.com> > >> An: "Ken.Peek@DiligentMinds.com" <Ken.Peek@DiligentMinds.com>, "Discussion of precise voltage measurement" <volt-nuts@febo.com> > >> Betreff: [volt-nuts] Bls: How accurate is a fluke after 9 years > >> > >> Hi All,What about Edward G Pettis resistor are they include in *THAT ARE PROPERLY WELDED* > >> > >> > >> > >> Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis: > >> > >> > >> IMHO: Vishay metal-foil resistors are very good, but *ONLY* if you get > >> them in a hermetic package. Depending on where you are in the world, > >> this can be rather expensive [more or less]. Precision wire-wound > >> resistors made with with low TCR Evanohm wire *THAT ARE PROPERLY WELDED* > >> are the next best thing, and they don't cost you your first born... > >> > >> > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > >> > >> > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
AM
Anton Moehammad
Mon, May 18, 2015 12:39 AM

Hi All,
in the Vishay web I found "a better in spec" resistor than  vha512, could somebody give a comment ?
http://www.vishaypg.com/docs/63120/hzseries.pdfThey also offer non standard value, anybody has experience in Vishay non standard value ?

Thank You

 Pada Senin, 18 Mei 2015 3:25, "acbern@gmx.de" <acbern@gmx.de> menulis:

hi andreas,

yes, it was a vha512 (maybe 516, but dont think so, dont have data handy). I ordered it from their berlin area based distributor (powertron). I ordered a couple, but different values, so it was single piece price. was some time ago, so price may have increased by now.

adrian

Gesendet: Sonntag, 17. Mai 2015 um 22:00 Uhr
Von: "Andreas Jahn" Andreas_-_Jahn@t-online.de
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Bls:  How accurate is a fluke after 9 years

Hello Adrian,

which actual type of resistor did you use.
And from where did you order it?
For which quantity is the 40 usd price? (single piece?)

I guess it was one of the hzseries of vishay. right?
Perhaps a Y4725 (VHA512Z?)
http://www.vishaypg.com/docs/63120/hzseries.pdf
Was it including PMO?

With best regards

Andreas

Am 17.05.2015 um 18:16 schrieb acbern@gmx.de:

I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper.
you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa.
I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5).

Adrian

Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr
Von: "Anton Moehammad via volt-nuts" volt-nuts@febo.com
An: "Ken.Peek@DiligentMinds.com" Ken.Peek@DiligentMinds.com, "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Bls:  How accurate is a fluke after 9 years

Hi All,What about Edward G Pettis resistor are they include in THAT ARE PROPERLY WELDED
 

      Pada Minggu, 17 Mei 2015 2:11, Ken Peek ken.peek@diligentminds.com menulis:
   

  IMHO:  Vishay metal-foil resistors are very good, but ONLY if you get
them in a hermetic package.  Depending on where you are in the world,
this can be rather expensive [more or less].  Precision wire-wound
resistors made with with low TCR Evanohm wire THAT ARE PROPERLY WELDED
are the next best thing, and they don't cost you your first born...


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

   


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi All, in the Vishay web I found "a better in spec" resistor than  vha512, could somebody give a comment ? http://www.vishaypg.com/docs/63120/hzseries.pdfThey also offer non standard value, anybody has experience in Vishay non standard value ? Thank You Pada Senin, 18 Mei 2015 3:25, "acbern@gmx.de" <acbern@gmx.de> menulis: hi andreas, yes, it was a vha512 (maybe 516, but dont think so, dont have data handy). I ordered it from their berlin area based distributor (powertron). I ordered a couple, but different values, so it was single piece price. was some time ago, so price may have increased by now. adrian > Gesendet: Sonntag, 17. Mai 2015 um 22:00 Uhr > Von: "Andreas Jahn" <Andreas_-_Jahn@t-online.de> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] Bls:  How accurate is a fluke after 9 years > > Hello Adrian, > > which actual type of resistor did you use. > And from where did you order it? > For which quantity is the 40 usd price? (single piece?) > > I guess it was one of the hzseries of vishay. right? > Perhaps a Y4725 (VHA512Z?) > http://www.vishaypg.com/docs/63120/hzseries.pdf > Was it including PMO? > > With best regards > > Andreas > > Am 17.05.2015 um 18:16 schrieb acbern@gmx.de: > > I have VPG hermetic foil resistors that drifted 0.2ppm/year (verified in cal lab). temp co. arround room temp <0,2ppm/k. my personal opinion, they are the best resistors for stability that you can purchse as a standard list item. Mine were < 40usd (4-wire, 0.1%, in reality they are better 0.01%). 2-wire types are probably cheaper. > > you cannot use standard foil resistors, they drift due to humidity, some 10 ppm easily pa. > > I would think Edward Petties do drift as well, most likely onsiderably more than the heremtic resistors, but probably less than vpg epoxy foil,(there is no drift spec against time from him), since his are not hermetic. And also, typically his temp drift is worse (1-3ppm/K versus <0,5). > > > > Adrian > > > >> Gesendet: Sonntag, 17. Mai 2015 um 03:40 Uhr > >> Von: "Anton Moehammad via volt-nuts" <volt-nuts@febo.com> > >> An: "Ken.Peek@DiligentMinds.com" <Ken.Peek@DiligentMinds.com>, "Discussion of precise voltage measurement" <volt-nuts@febo.com> > >> Betreff: [volt-nuts] Bls:  How accurate is a fluke after 9 years > >> > >> Hi All,What about Edward G Pettis resistor are they include in *THAT ARE PROPERLY WELDED* > >>  > >> > >> > >>      Pada Minggu, 17 Mei 2015 2:11, Ken Peek <ken.peek@diligentminds.com> menulis: > >>    > >> > >>  IMHO:  Vishay metal-foil resistors are very good, but *ONLY* if you get > >> them in a hermetic package.  Depending on where you are in the world, > >> this can be rather expensive [more or less].  Precision wire-wound > >> resistors made with with low TCR Evanohm wire *THAT ARE PROPERLY WELDED* > >> are the next best thing, and they don't cost you your first born... > >> > >> > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > >> > >>    > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
A
acbern@gmx.de
Mon, May 18, 2015 6:13 AM

it is this resistor, in its z-foil version.
they can be ordered from vpg distributors also in small qty (some may have a min. order value), lead time is usually pretty high.

Gesendet: Montag, 18. Mai 2015 um 02:39 Uhr
Von: "Anton Moehammad via volt-nuts" volt-nuts@febo.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Bls:  Bls:  How accurate is a fluke after 9 years

Hi All,
in the Vishay web I found "a better in spec" resistor than  vha512, could somebody give a comment ?
http://www.vishaypg.com/docs/63120/hzseries.pdfThey also offer non standard value, anybody has experience in Vishay non standard value ?

Thank You

it is this resistor, in its z-foil version. they can be ordered from vpg distributors also in small qty (some may have a min. order value), lead time is usually pretty high. > Gesendet: Montag, 18. Mai 2015 um 02:39 Uhr > Von: "Anton Moehammad via volt-nuts" <volt-nuts@febo.com> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: [volt-nuts] Bls: Bls: How accurate is a fluke after 9 years > > Hi All, > in the Vishay web I found "a better in spec" resistor than  vha512, could somebody give a comment ? > http://www.vishaypg.com/docs/63120/hzseries.pdfThey also offer non standard value, anybody has experience in Vishay non standard value ? > > Thank You > >
TM
Todd Micallef
Mon, May 18, 2015 8:28 PM

The 5700A5720A verification test lists a maximum deviation from nominal on
the resistance ranges. Is there any information on how well Fluke's hybrid
resistors are maintaining their specs over the lifetime of the calibrator?
Anyone have access to long term records of the resistance values?

Todd

On Mon, May 18, 2015 at 2:13 AM, acbern@gmx.de wrote:

it is this resistor, in its z-foil version.
they can be ordered from vpg distributors also in small qty (some may have
a min. order value), lead time is usually pretty high.

Gesendet: Montag, 18. Mai 2015 um 02:39 Uhr
Von: "Anton Moehammad via volt-nuts" volt-nuts@febo.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Bls:  Bls:  How accurate is a fluke after 9 years

Hi All,
in the Vishay web I found "a better in spec" resistor than  vha512,

could somebody give a comment ?

standard value, anybody has experience in Vishay non standard value ?

Thank You


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The 5700A5720A verification test lists a maximum deviation from nominal on the resistance ranges. Is there any information on how well Fluke's hybrid resistors are maintaining their specs over the lifetime of the calibrator? Anyone have access to long term records of the resistance values? Todd On Mon, May 18, 2015 at 2:13 AM, <acbern@gmx.de> wrote: > it is this resistor, in its z-foil version. > they can be ordered from vpg distributors also in small qty (some may have > a min. order value), lead time is usually pretty high. > > > > Gesendet: Montag, 18. Mai 2015 um 02:39 Uhr > > Von: "Anton Moehammad via volt-nuts" <volt-nuts@febo.com> > > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > Betreff: [volt-nuts] Bls: Bls: How accurate is a fluke after 9 years > > > > Hi All, > > in the Vishay web I found "a better in spec" resistor than vha512, > could somebody give a comment ? > > http://www.vishaypg.com/docs/63120/hzseries.pdfThey also offer non > standard value, anybody has experience in Vishay non standard value ? > > > > Thank You > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >