EB
ed breya
Fri, Nov 30, 2012 4:00 AM
I just junked out a very beat up old Fluke 803 differential
voltmeter, and found deep within, an old-school Cd/Hg standard cell.
It was well protected in an aluminum box, and wrapped in foam and
foil. It looks brand-new, and still measures around 1.018... V. I'd
like to keep this one as another reference point if it's still good.
I assume that it just wasn't used much, or that the Fluke circuits
were very good at not loading it down.
I'm sure it is the original unit installed in the instrument - marked
5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
any other info but a serial number.
So, I'm wondering if a 52 year old standard cell can still be OK, and
if anyone knows the specs on these, or where to find the info. I
don't know if it's possible, but I'd like to find what the official
voltage was supposed to be to a few more digits resolution. I think
various types and brands each had slightly different nominal voltages
around that determined by the basic chemistry. I remember in the old
days, every one I saw included a sticker with the 25 deg C exact
voltage measured as accurately as possible back then against the NBS.
I'd like to especially know if this is a saturated or unsaturated cell type.
Ed
I just junked out a very beat up old Fluke 803 differential
voltmeter, and found deep within, an old-school Cd/Hg standard cell.
It was well protected in an aluminum box, and wrapped in foam and
foil. It looks brand-new, and still measures around 1.018... V. I'd
like to keep this one as another reference point if it's still good.
I assume that it just wasn't used much, or that the Fluke circuits
were very good at not loading it down.
I'm sure it is the original unit installed in the instrument - marked
5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
any other info but a serial number.
So, I'm wondering if a 52 year old standard cell can still be OK, and
if anyone knows the specs on these, or where to find the info. I
don't know if it's possible, but I'd like to find what the official
voltage was supposed to be to a few more digits resolution. I think
various types and brands each had slightly different nominal voltages
around that determined by the basic chemistry. I remember in the old
days, every one I saw included a sticker with the 25 deg C exact
voltage measured as accurately as possible back then against the NBS.
I'd like to especially know if this is a saturated or unsaturated cell type.
Ed
W
WB6BNQ
Fri, Nov 30, 2012 7:15 AM
Hi Ed,
The type of cell you have is an UNsaturated type. These types are portable. The
saturated type of cells are non-portable and very temperature sensitive requiring
the use of a constant temperature enclousure.
Here is a web site that describes these devices that is quite good:
http://conradhoffman.com/stdcell.htm
Bill....WB6BNQ
ed breya wrote:
I just junked out a very beat up old Fluke 803 differential
voltmeter, and found deep within, an old-school Cd/Hg standard cell.
It was well protected in an aluminum box, and wrapped in foam and
foil. It looks brand-new, and still measures around 1.018... V. I'd
like to keep this one as another reference point if it's still good.
I assume that it just wasn't used much, or that the Fluke circuits
were very good at not loading it down.
I'm sure it is the original unit installed in the instrument - marked
5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
any other info but a serial number.
So, I'm wondering if a 52 year old standard cell can still be OK, and
if anyone knows the specs on these, or where to find the info. I
don't know if it's possible, but I'd like to find what the official
voltage was supposed to be to a few more digits resolution. I think
various types and brands each had slightly different nominal voltages
around that determined by the basic chemistry. I remember in the old
days, every one I saw included a sticker with the 25 deg C exact
voltage measured as accurately as possible back then against the NBS.
I'd like to especially know if this is a saturated or unsaturated cell type.
Ed
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi Ed,
The type of cell you have is an UNsaturated type. These types are portable. The
saturated type of cells are non-portable and very temperature sensitive requiring
the use of a constant temperature enclousure.
Here is a web site that describes these devices that is quite good:
http://conradhoffman.com/stdcell.htm
Bill....WB6BNQ
ed breya wrote:
> I just junked out a very beat up old Fluke 803 differential
> voltmeter, and found deep within, an old-school Cd/Hg standard cell.
> It was well protected in an aluminum box, and wrapped in foam and
> foil. It looks brand-new, and still measures around 1.018... V. I'd
> like to keep this one as another reference point if it's still good.
> I assume that it just wasn't used much, or that the Fluke circuits
> were very good at not loading it down.
>
> I'm sure it is the original unit installed in the instrument - marked
> 5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
> any other info but a serial number.
>
> So, I'm wondering if a 52 year old standard cell can still be OK, and
> if anyone knows the specs on these, or where to find the info. I
> don't know if it's possible, but I'd like to find what the official
> voltage was supposed to be to a few more digits resolution. I think
> various types and brands each had slightly different nominal voltages
> around that determined by the basic chemistry. I remember in the old
> days, every one I saw included a sticker with the 25 deg C exact
> voltage measured as accurately as possible back then against the NBS.
> I'd like to especially know if this is a saturated or unsaturated cell type.
>
> Ed
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
FS
Fred Schneider
Fri, Nov 30, 2012 7:31 AM
http://www.pa4tim.nl/?p=2141. My saturated cells, so you can compare them. I think you have unsaturated ones because saturated ones can not be transported and must be kept in an oven.
Most unsaturated cells have a limited lifetime, i read around 10-20 years.
Fred PA4TIM
Op 30 nov. 2012 om 08:15 heeft WB6BNQ wb6bnq@cox.net het volgende geschreven:
Hi Ed,
The type of cell you have is an UNsaturated type. These types are portable. The
saturated type of cells are non-portable and very temperature sensitive requiring
the use of a constant temperature enclousure.
Here is a web site that describes these devices that is quite good:
http://conradhoffman.com/stdcell.htm
Bill....WB6BNQ
ed breya wrote:
I just junked out a very beat up old Fluke 803 differential
voltmeter, and found deep within, an old-school Cd/Hg standard cell.
It was well protected in an aluminum box, and wrapped in foam and
foil. It looks brand-new, and still measures around 1.018... V. I'd
like to keep this one as another reference point if it's still good.
I assume that it just wasn't used much, or that the Fluke circuits
were very good at not loading it down.
I'm sure it is the original unit installed in the instrument - marked
5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
any other info but a serial number.
So, I'm wondering if a 52 year old standard cell can still be OK, and
if anyone knows the specs on these, or where to find the info. I
don't know if it's possible, but I'd like to find what the official
voltage was supposed to be to a few more digits resolution. I think
various types and brands each had slightly different nominal voltages
around that determined by the basic chemistry. I remember in the old
days, every one I saw included a sticker with the 25 deg C exact
voltage measured as accurately as possible back then against the NBS.
I'd like to especially know if this is a saturated or unsaturated cell type.
Ed
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
http://www.pa4tim.nl/?p=2141. My saturated cells, so you can compare them. I think you have unsaturated ones because saturated ones can not be transported and must be kept in an oven.
Most unsaturated cells have a limited lifetime, i read around 10-20 years.
Fred PA4TIM
Op 30 nov. 2012 om 08:15 heeft WB6BNQ <wb6bnq@cox.net> het volgende geschreven:
> Hi Ed,
>
> The type of cell you have is an UNsaturated type. These types are portable. The
> saturated type of cells are non-portable and very temperature sensitive requiring
> the use of a constant temperature enclousure.
>
> Here is a web site that describes these devices that is quite good:
>
> http://conradhoffman.com/stdcell.htm
>
>
> Bill....WB6BNQ
>
>
>
> ed breya wrote:
>
>> I just junked out a very beat up old Fluke 803 differential
>> voltmeter, and found deep within, an old-school Cd/Hg standard cell.
>> It was well protected in an aluminum box, and wrapped in foam and
>> foil. It looks brand-new, and still measures around 1.018... V. I'd
>> like to keep this one as another reference point if it's still good.
>> I assume that it just wasn't used much, or that the Fluke circuits
>> were very good at not loading it down.
>>
>> I'm sure it is the original unit installed in the instrument - marked
>> 5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
>> any other info but a serial number.
>>
>> So, I'm wondering if a 52 year old standard cell can still be OK, and
>> if anyone knows the specs on these, or where to find the info. I
>> don't know if it's possible, but I'd like to find what the official
>> voltage was supposed to be to a few more digits resolution. I think
>> various types and brands each had slightly different nominal voltages
>> around that determined by the basic chemistry. I remember in the old
>> days, every one I saw included a sticker with the 25 deg C exact
>> voltage measured as accurately as possible back then against the NBS.
>> I'd like to especially know if this is a saturated or unsaturated cell type.
>>
>> Ed
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
MK
m k
Fri, Nov 30, 2012 7:33 AM
Hi,
Well from that site the tempco is about 5 microvolts per degree, and from the bottom of the page:
The un-saturated cell should have an e.m.f. of 1.0190 volts @ 20°C when new. When the voltage drops to 1.0183 volts, the
cell should be discarded, as it is no longer stable. Most reputable labs will not certify a cell below this value.
The NBS recorded data on approximately 600 un-saturated cells. About 5% showed an increase of e.m.f. with time, the average
change being 28 µV/year. The remaining 95% decrease in e.m.f., at an average of 85 µV/year. Of that group, nearly half
changed by more than 50 µV/year, and one fourth by more than 100 µV/year. That data suggests that certification at
yearly intervals is required to insure .01% accuracy. The expected life of un-saturated cells is 7-14 years.
So my guess would be that it is near the end of its life, but measure and see.
MK
Date: Thu, 29 Nov 2012 23:15:19 -0800
From: wb6bnq@cox.net
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] How long can standard cells last?
Hi Ed,
The type of cell you have is an UNsaturated type. These types are portable. The
saturated type of cells are non-portable and very temperature sensitive requiring
the use of a constant temperature enclousure.
Here is a web site that describes these devices that is quite good:
http://conradhoffman.com/stdcell.htm
Bill....WB6BNQ
ed breya wrote:
I just junked out a very beat up old Fluke 803 differential
voltmeter, and found deep within, an old-school Cd/Hg standard cell.
It was well protected in an aluminum box, and wrapped in foam and
foil. It looks brand-new, and still measures around 1.018... V. I'd
like to keep this one as another reference point if it's still good.
I assume that it just wasn't used much, or that the Fluke circuits
were very good at not loading it down.
I'm sure it is the original unit installed in the instrument - marked
5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
any other info but a serial number.
So, I'm wondering if a 52 year old standard cell can still be OK, and
if anyone knows the specs on these, or where to find the info. I
don't know if it's possible, but I'd like to find what the official
voltage was supposed to be to a few more digits resolution. I think
various types and brands each had slightly different nominal voltages
around that determined by the basic chemistry. I remember in the old
days, every one I saw included a sticker with the 25 deg C exact
voltage measured as accurately as possible back then against the NBS.
I'd like to especially know if this is a saturated or unsaturated cell type.
Ed
Hi,
Well from that site the tempco is about 5 microvolts per degree, and from the bottom of the page:
The un-saturated cell should have an e.m.f. of 1.0190 volts @ 20°C when new. When the voltage drops to 1.0183 volts, the
cell should be discarded, as it is no longer stable. Most reputable labs will not certify a cell below this value.
The NBS recorded data on approximately 600 un-saturated cells. About 5% showed an increase of e.m.f. with time, the average
change being 28 µV/year. The remaining 95% decrease in e.m.f., at an average of 85 µV/year. Of that group, nearly half
changed by more than 50 µV/year, and one fourth by more than 100 µV/year. That data suggests that certification at
yearly intervals is required to insure .01% accuracy. The expected life of un-saturated cells is 7-14 years.
So my guess would be that it is near the end of its life, but measure and see.
MK
> Date: Thu, 29 Nov 2012 23:15:19 -0800
> From: wb6bnq@cox.net
> To: volt-nuts@febo.com
> Subject: Re: [volt-nuts] How long can standard cells last?
>
> Hi Ed,
>
> The type of cell you have is an UNsaturated type. These types are portable. The
> saturated type of cells are non-portable and very temperature sensitive requiring
> the use of a constant temperature enclousure.
>
> Here is a web site that describes these devices that is quite good:
>
> http://conradhoffman.com/stdcell.htm
>
>
> Bill....WB6BNQ
>
>
>
> ed breya wrote:
>
> > I just junked out a very beat up old Fluke 803 differential
> > voltmeter, and found deep within, an old-school Cd/Hg standard cell.
> > It was well protected in an aluminum box, and wrapped in foam and
> > foil. It looks brand-new, and still measures around 1.018... V. I'd
> > like to keep this one as another reference point if it's still good.
> > I assume that it just wasn't used much, or that the Fluke circuits
> > were very good at not loading it down.
> >
> > I'm sure it is the original unit installed in the instrument - marked
> > 5/12/1960. It is a Muirhead D-845-C. There's no test voltage tag or
> > any other info but a serial number.
> >
> > So, I'm wondering if a 52 year old standard cell can still be OK, and
> > if anyone knows the specs on these, or where to find the info. I
> > don't know if it's possible, but I'd like to find what the official
> > voltage was supposed to be to a few more digits resolution. I think
> > various types and brands each had slightly different nominal voltages
> > around that determined by the basic chemistry. I remember in the old
> > days, every one I saw included a sticker with the 25 deg C exact
> > voltage measured as accurately as possible back then against the NBS.
> > I'd like to especially know if this is a saturated or unsaturated cell type.
> >
> > Ed
> >