What's All This Low Thermal EMF Test Lead Stuff?

SK
Stan Katz
Tue, Aug 19, 2014 6:07 AM

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my instrumentation
pecking order.  I have all the necessary gear to calibrate these DMMs
according to ancient HP documentation. At the top of my cal. chain is the
731B, called out in HP3456 original documentation.  This hierarchy places
me in the near-beer, or junior member status of the group. I look forward
to being educated, and/or corrected on my understanding of the use of test
leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use as
thick a solid lead of pure copper wire as you can find, and insert the wire
into the drill hole on the banana terminal. If the copper is pure, and has
been properly cleaned, the thermal emf's on both identical length leads
should all be balanced, and cancel out. In any case, pure copper-copper
connections generate the lowest thermal emf. I  will agree that manhandling
16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over copper
stereo banana plugs for my connections ( I deal in low voltage work
exclusively), as well as gold plated spade lugs to go under the banana
screw-downs. ( My budget ruled out gold over beryllium copper Pomona brand
spades. )  I will then experiment between the two connector types. As for
connections, it seems to me the best course is just to screw down the
banana plugs, or in the case of the spades, just crimp. I'll wing it on the
crimping, and see if simple tools can perform adequately. I would avoid
solder, since how can one form identical topological spots of solder on
every connection, deposited at the exact same place on each connection, and
ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I think
solder is out. Of course, if one wishes to risk one's health, and that of
the family, one can track down a source of cadmium, and mix up a lot of low
thermal emf solder (cadmium solder is banned in the US, and EU)...not for
me.

Don't worry, I won't be using tin plated wire, I managed to pick up a spool
of silver plated 16 gauge wire for my investigations.  Am I on some other
planet for choosing the balanced thermal emf approach for precision test
lead applications, as opposed to going to great lengths to eliminate every
trace of emf?

Please forgive me if these emf discussions have been between engineers
working outdoors in the oil fields of North Dakota, or out on a North Sea
drilling platform. In that case, go for all the super low emf techniques
you can lay your hands on.

Good Luck,
Stan

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an HP3458. I do own HP3456 DMMs that are at the top of my instrumentation pecking order. I have all the necessary gear to calibrate these DMMs according to ancient HP documentation. At the top of my cal. chain is the 731B, called out in HP3456 original documentation. This hierarchy places me in the near-beer, or junior member status of the group. I look forward to being educated, and/or corrected on my understanding of the use of test leads with precision instrumentation. I don't find much ancient HP documentation on test leads. The only recommendations in the era of the 3456 back to the 1960's is to use as thick a solid lead of pure copper wire as you can find, and insert the wire into the drill hole on the banana terminal. If the copper is pure, and has been properly cleaned, the thermal emf's on both identical length leads should all be balanced, and cancel out. In any case, pure copper-copper connections generate the lowest thermal emf. I will agree that manhandling 16 gauge solid wire can be very inconvenient. The path I have taken recently is to order Nakamichi gold over copper stereo banana plugs for my connections ( I deal in low voltage work exclusively), as well as gold plated spade lugs to go under the banana screw-downs. ( My budget ruled out gold over beryllium copper Pomona brand spades. ) I will then experiment between the two connector types. As for connections, it seems to me the best course is just to screw down the banana plugs, or in the case of the spades, just crimp. I'll wing it on the crimping, and see if simple tools can perform adequately. I would avoid solder, since how can one form identical topological spots of solder on every connection, deposited at the exact same place on each connection, and ensure the exact same weight of solder, to the microgram, on each connection. Since my modus operandi is to aim for balanced emfs, I think solder is out. Of course, if one wishes to risk one's health, and that of the family, one can track down a source of cadmium, and mix up a lot of low thermal emf solder (cadmium solder is banned in the US, and EU)...not for me. Don't worry, I won't be using tin plated wire, I managed to pick up a spool of silver plated 16 gauge wire for my investigations. Am I on some other planet for choosing the balanced thermal emf approach for precision test lead applications, as opposed to going to great lengths to eliminate every trace of emf? Please forgive me if these emf discussions have been between engineers working outdoors in the oil fields of North Dakota, or out on a North Sea drilling platform. In that case, go for all the super low emf techniques you can lay your hands on. Good Luck, Stan
A
acbern@gmx.de
Tue, Aug 19, 2014 7:37 AM

generally I think, and at least for your gear, if you use copper plugs of some type and even solder the wires, thats perfectly fine.
also for a 3458a, and after proper temp stabilization after connecting the cables (5min) which you always need, these are fine, have not measrured any difference to crimped cables. also the type of cable plating (tin/silver..) does not make much difference.
there were many discussions over time, but all the cadmium solder and non plated copper type debates are not making much sense down to the 3458a level.
I have not done comparison tests with nanovolt meters, as mine have fixed cables, so no option to change them without destroying them, but note that e.g. the 182 nanovolt meter has a mil type plug (no special contacts). sources of error are also things like where to connect the guard to, these often make a much bigger difference (I have seen these many times, causing voltage differences of several uV depending where in the setup you connected the guard, while otherwise the measurement was perfectly stable), and I bet while people put a lot of effort into the emf discussion, the guard issue is more often causing errors in measurements in real life. and thats just one example of additional causes of errors.
cheers

Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr
Von: "Stan Katz" stan.katz.hk@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my instrumentation
pecking order.  I have all the necessary gear to calibrate these DMMs
according to ancient HP documentation. At the top of my cal. chain is the
731B, called out in HP3456 original documentation.  This hierarchy places
me in the near-beer, or junior member status of the group. I look forward
to being educated, and/or corrected on my understanding of the use of test
leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use as
thick a solid lead of pure copper wire as you can find, and insert the wire
into the drill hole on the banana terminal. If the copper is pure, and has
been properly cleaned, the thermal emf's on both identical length leads
should all be balanced, and cancel out. In any case, pure copper-copper
connections generate the lowest thermal emf. I  will agree that manhandling
16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over copper
stereo banana plugs for my connections ( I deal in low voltage work
exclusively), as well as gold plated spade lugs to go under the banana
screw-downs. ( My budget ruled out gold over beryllium copper Pomona brand
spades. )  I will then experiment between the two connector types. As for
connections, it seems to me the best course is just to screw down the
banana plugs, or in the case of the spades, just crimp. I'll wing it on the
crimping, and see if simple tools can perform adequately. I would avoid
solder, since how can one form identical topological spots of solder on
every connection, deposited at the exact same place on each connection, and
ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I think
solder is out. Of course, if one wishes to risk one's health, and that of
the family, one can track down a source of cadmium, and mix up a lot of low
thermal emf solder (cadmium solder is banned in the US, and EU)...not for
me.

Don't worry, I won't be using tin plated wire, I managed to pick up a spool
of silver plated 16 gauge wire for my investigations.  Am I on some other
planet for choosing the balanced thermal emf approach for precision test
lead applications, as opposed to going to great lengths to eliminate every
trace of emf?

Please forgive me if these emf discussions have been between engineers
working outdoors in the oil fields of North Dakota, or out on a North Sea
drilling platform. In that case, go for all the super low emf techniques
you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

generally I think, and at least for your gear, if you use copper plugs of some type and even solder the wires, thats perfectly fine. also for a 3458a, and after proper temp stabilization after connecting the cables (5min) which you always need, these are fine, have not measrured any difference to crimped cables. also the type of cable plating (tin/silver..) does not make much difference. there were many discussions over time, but all the cadmium solder and non plated copper type debates are not making much sense down to the 3458a level. I have not done comparison tests with nanovolt meters, as mine have fixed cables, so no option to change them without destroying them, but note that e.g. the 182 nanovolt meter has a mil type plug (no special contacts). sources of error are also things like where to connect the guard to, these often make a much bigger difference (I have seen these many times, causing voltage differences of several uV depending where in the setup you connected the guard, while otherwise the measurement was perfectly stable), and I bet while people put a lot of effort into the emf discussion, the guard issue is more often causing errors in measurements in real life. and thats just one example of additional causes of errors. cheers > Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr > Von: "Stan Katz" <stan.katz.hk@gmail.com> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? > > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > HP3458. I do own HP3456 DMMs that are at the top of my instrumentation > pecking order. I have all the necessary gear to calibrate these DMMs > according to ancient HP documentation. At the top of my cal. chain is the > 731B, called out in HP3456 original documentation. This hierarchy places > me in the near-beer, or junior member status of the group. I look forward > to being educated, and/or corrected on my understanding of the use of test > leads with precision instrumentation. > > I don't find much ancient HP documentation on test leads. The only > recommendations in the era of the 3456 back to the 1960's is to use as > thick a solid lead of pure copper wire as you can find, and insert the wire > into the drill hole on the banana terminal. If the copper is pure, and has > been properly cleaned, the thermal emf's on both identical length leads > should all be balanced, and cancel out. In any case, pure copper-copper > connections generate the lowest thermal emf. I will agree that manhandling > 16 gauge solid wire can be very inconvenient. > > The path I have taken recently is to order Nakamichi gold over copper > stereo banana plugs for my connections ( I deal in low voltage work > exclusively), as well as gold plated spade lugs to go under the banana > screw-downs. ( My budget ruled out gold over beryllium copper Pomona brand > spades. ) I will then experiment between the two connector types. As for > connections, it seems to me the best course is just to screw down the > banana plugs, or in the case of the spades, just crimp. I'll wing it on the > crimping, and see if simple tools can perform adequately. I would avoid > solder, since how can one form identical topological spots of solder on > every connection, deposited at the exact same place on each connection, and > ensure the exact same weight of solder, to the microgram, on each > connection. Since my modus operandi is to aim for balanced emfs, I think > solder is out. Of course, if one wishes to risk one's health, and that of > the family, one can track down a source of cadmium, and mix up a lot of low > thermal emf solder (cadmium solder is banned in the US, and EU)...not for > me. > > Don't worry, I won't be using tin plated wire, I managed to pick up a spool > of silver plated 16 gauge wire for my investigations. Am I on some other > planet for choosing the balanced thermal emf approach for precision test > lead applications, as opposed to going to great lengths to eliminate every > trace of emf? > > Please forgive me if these emf discussions have been between engineers > working outdoors in the oil fields of North Dakota, or out on a North Sea > drilling platform. In that case, go for all the super low emf techniques > you can lay your hands on. > > Good Luck, > Stan > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
TM
Todd Micallef
Tue, Aug 19, 2014 8:12 PM

Stan,

I recently picked up some of these ...
http://www.douglasconnection.com/Furez-TSTWP30NP-Bare-Copper-Banana-Plug-Connectors-Pair-FZTSTWP30NP.htm

They are a little pricey and are made for 12ga wire. I think they are OK
for semi-permanent use. A lot of use will probably scratch the soft metal.
I plan on trying them with my Keithley 181 plugged into a low thermal
scanner.
The original Keithley cable will be difficult to terminate. I may have to
go with crimped spade lugs.

Also, I am all for joining a beer nuts group.

Todd

On Tue, Aug 19, 2014 at 2:07 AM, Stan Katz stan.katz.hk@gmail.com wrote:

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my instrumentation
pecking order.  I have all the necessary gear to calibrate these DMMs
according to ancient HP documentation. At the top of my cal. chain is the
731B, called out in HP3456 original documentation.  This hierarchy places
me in the near-beer, or junior member status of the group. I look forward
to being educated, and/or corrected on my understanding of the use of test
leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use as
thick a solid lead of pure copper wire as you can find, and insert the wire
into the drill hole on the banana terminal. If the copper is pure, and has
been properly cleaned, the thermal emf's on both identical length leads
should all be balanced, and cancel out. In any case, pure copper-copper
connections generate the lowest thermal emf. I  will agree that manhandling
16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over copper
stereo banana plugs for my connections ( I deal in low voltage work
exclusively), as well as gold plated spade lugs to go under the banana
screw-downs. ( My budget ruled out gold over beryllium copper Pomona brand
spades. )  I will then experiment between the two connector types. As for
connections, it seems to me the best course is just to screw down the
banana plugs, or in the case of the spades, just crimp. I'll wing it on the
crimping, and see if simple tools can perform adequately. I would avoid
solder, since how can one form identical topological spots of solder on
every connection, deposited at the exact same place on each connection, and
ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I think
solder is out. Of course, if one wishes to risk one's health, and that of
the family, one can track down a source of cadmium, and mix up a lot of low
thermal emf solder (cadmium solder is banned in the US, and EU)...not for
me.

Don't worry, I won't be using tin plated wire, I managed to pick up a spool
of silver plated 16 gauge wire for my investigations.  Am I on some other
planet for choosing the balanced thermal emf approach for precision test
lead applications, as opposed to going to great lengths to eliminate every
trace of emf?

Please forgive me if these emf discussions have been between engineers
working outdoors in the oil fields of North Dakota, or out on a North Sea
drilling platform. In that case, go for all the super low emf techniques
you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Stan, I recently picked up some of these ... http://www.douglasconnection.com/Furez-TSTWP30NP-Bare-Copper-Banana-Plug-Connectors-Pair-FZTSTWP30NP.htm They are a little pricey and are made for 12ga wire. I think they are OK for semi-permanent use. A lot of use will probably scratch the soft metal. I plan on trying them with my Keithley 181 plugged into a low thermal scanner. The original Keithley cable will be difficult to terminate. I may have to go with crimped spade lugs. Also, I am all for joining a beer nuts group. Todd On Tue, Aug 19, 2014 at 2:07 AM, Stan Katz <stan.katz.hk@gmail.com> wrote: > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > HP3458. I do own HP3456 DMMs that are at the top of my instrumentation > pecking order. I have all the necessary gear to calibrate these DMMs > according to ancient HP documentation. At the top of my cal. chain is the > 731B, called out in HP3456 original documentation. This hierarchy places > me in the near-beer, or junior member status of the group. I look forward > to being educated, and/or corrected on my understanding of the use of test > leads with precision instrumentation. > > I don't find much ancient HP documentation on test leads. The only > recommendations in the era of the 3456 back to the 1960's is to use as > thick a solid lead of pure copper wire as you can find, and insert the wire > into the drill hole on the banana terminal. If the copper is pure, and has > been properly cleaned, the thermal emf's on both identical length leads > should all be balanced, and cancel out. In any case, pure copper-copper > connections generate the lowest thermal emf. I will agree that manhandling > 16 gauge solid wire can be very inconvenient. > > The path I have taken recently is to order Nakamichi gold over copper > stereo banana plugs for my connections ( I deal in low voltage work > exclusively), as well as gold plated spade lugs to go under the banana > screw-downs. ( My budget ruled out gold over beryllium copper Pomona brand > spades. ) I will then experiment between the two connector types. As for > connections, it seems to me the best course is just to screw down the > banana plugs, or in the case of the spades, just crimp. I'll wing it on the > crimping, and see if simple tools can perform adequately. I would avoid > solder, since how can one form identical topological spots of solder on > every connection, deposited at the exact same place on each connection, and > ensure the exact same weight of solder, to the microgram, on each > connection. Since my modus operandi is to aim for balanced emfs, I think > solder is out. Of course, if one wishes to risk one's health, and that of > the family, one can track down a source of cadmium, and mix up a lot of low > thermal emf solder (cadmium solder is banned in the US, and EU)...not for > me. > > Don't worry, I won't be using tin plated wire, I managed to pick up a spool > of silver plated 16 gauge wire for my investigations. Am I on some other > planet for choosing the balanced thermal emf approach for precision test > lead applications, as opposed to going to great lengths to eliminate every > trace of emf? > > Please forgive me if these emf discussions have been between engineers > working outdoors in the oil fields of North Dakota, or out on a North Sea > drilling platform. In that case, go for all the super low emf techniques > you can lay your hands on. > > Good Luck, > Stan > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
SK
Stan Katz
Tue, Aug 19, 2014 11:35 PM

You are on the money that guarding/shielding are sadly neglected as
subjects, in their own right. Posts in this group consist primarily of
reports of guard leakage in member instruments, and repair details.
Discussion of proper lash ups of cal. instruments to home built devices, or
even work prototypes, would be very welcome by me. I have read Analog
Devices AN-347, but a tutorial with problems that the student must complete
the lash ups for, would give me a better idea if I've grasped the subject
matter. There are two text books mentioned at the end of AN-347, one by
Morrison, the other by Ott. Any suggestion as to which to buy? The latest
texts by these authors are pricey, so that's why I'd like to pick just one.

On Tue, Aug 19, 2014 at 3:37 AM, acbern@gmx.de wrote:

generally I think, and at least for your gear, if you use copper plugs of
some type and even solder the wires, thats perfectly fine.
also for a 3458a, and after proper temp stabilization after connecting the
cables (5min) which you always need, these are fine, have not measrured any
difference to crimped cables. also the type of cable plating (tin/silver..)
does not make much difference.
there were many discussions over time, but all the cadmium solder and non
plated copper type debates are not making much sense down to the 3458a
level.
I have not done comparison tests with nanovolt meters, as mine have fixed
cables, so no option to change them without destroying them, but note that
e.g. the 182 nanovolt meter has a mil type plug (no special contacts).
sources of error are also things like where to connect the guard to, these
often make a much bigger difference (I have seen these many times, causing
voltage differences of several uV depending where in the setup you
connected the guard, while otherwise the measurement was perfectly stable),
and I bet while people put a lot of effort into the emf discussion, the
guard issue is more often causing errors in measurements in real life. and
thats just one example of additional causes of errors.
cheers

Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr
Von: "Stan Katz" stan.katz.hk@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my instrumentation
pecking order.  I have all the necessary gear to calibrate these DMMs
according to ancient HP documentation. At the top of my cal. chain is the
731B, called out in HP3456 original documentation.  This hierarchy places
me in the near-beer, or junior member status of the group. I look forward
to being educated, and/or corrected on my understanding of the use of

test

leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use as
thick a solid lead of pure copper wire as you can find, and insert the

wire

into the drill hole on the banana terminal. If the copper is pure, and

has

been properly cleaned, the thermal emf's on both identical length leads
should all be balanced, and cancel out. In any case, pure copper-copper
connections generate the lowest thermal emf. I  will agree that

manhandling

16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over copper
stereo banana plugs for my connections ( I deal in low voltage work
exclusively), as well as gold plated spade lugs to go under the banana
screw-downs. ( My budget ruled out gold over beryllium copper Pomona

brand

spades. )  I will then experiment between the two connector types. As for
connections, it seems to me the best course is just to screw down the
banana plugs, or in the case of the spades, just crimp. I'll wing it on

the

crimping, and see if simple tools can perform adequately. I would avoid
solder, since how can one form identical topological spots of solder on
every connection, deposited at the exact same place on each connection,

and

ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I think
solder is out. Of course, if one wishes to risk one's health, and that of
the family, one can track down a source of cadmium, and mix up a lot of

low

thermal emf solder (cadmium solder is banned in the US, and EU)...not for
me.

Don't worry, I won't be using tin plated wire, I managed to pick up a

spool

of silver plated 16 gauge wire for my investigations.  Am I on some other
planet for choosing the balanced thermal emf approach for precision test
lead applications, as opposed to going to great lengths to eliminate

every

trace of emf?

Please forgive me if these emf discussions have been between engineers
working outdoors in the oil fields of North Dakota, or out on a North Sea
drilling platform. In that case, go for all the super low emf techniques
you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

You are on the money that guarding/shielding are sadly neglected as subjects, in their own right. Posts in this group consist primarily of reports of guard leakage in member instruments, and repair details. Discussion of proper lash ups of cal. instruments to home built devices, or even work prototypes, would be very welcome by me. I have read Analog Devices AN-347, but a tutorial with problems that the student must complete the lash ups for, would give me a better idea if I've grasped the subject matter. There are two text books mentioned at the end of AN-347, one by Morrison, the other by Ott. Any suggestion as to which to buy? The latest texts by these authors are pricey, so that's why I'd like to pick just one. On Tue, Aug 19, 2014 at 3:37 AM, <acbern@gmx.de> wrote: > generally I think, and at least for your gear, if you use copper plugs of > some type and even solder the wires, thats perfectly fine. > also for a 3458a, and after proper temp stabilization after connecting the > cables (5min) which you always need, these are fine, have not measrured any > difference to crimped cables. also the type of cable plating (tin/silver..) > does not make much difference. > there were many discussions over time, but all the cadmium solder and non > plated copper type debates are not making much sense down to the 3458a > level. > I have not done comparison tests with nanovolt meters, as mine have fixed > cables, so no option to change them without destroying them, but note that > e.g. the 182 nanovolt meter has a mil type plug (no special contacts). > sources of error are also things like where to connect the guard to, these > often make a much bigger difference (I have seen these many times, causing > voltage differences of several uV depending where in the setup you > connected the guard, while otherwise the measurement was perfectly stable), > and I bet while people put a lot of effort into the emf discussion, the > guard issue is more often causing errors in measurements in real life. and > thats just one example of additional causes of errors. > cheers > > > Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr > > Von: "Stan Katz" <stan.katz.hk@gmail.com> > > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? > > > > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > > HP3458. I do own HP3456 DMMs that are at the top of my instrumentation > > pecking order. I have all the necessary gear to calibrate these DMMs > > according to ancient HP documentation. At the top of my cal. chain is the > > 731B, called out in HP3456 original documentation. This hierarchy places > > me in the near-beer, or junior member status of the group. I look forward > > to being educated, and/or corrected on my understanding of the use of > test > > leads with precision instrumentation. > > > > I don't find much ancient HP documentation on test leads. The only > > recommendations in the era of the 3456 back to the 1960's is to use as > > thick a solid lead of pure copper wire as you can find, and insert the > wire > > into the drill hole on the banana terminal. If the copper is pure, and > has > > been properly cleaned, the thermal emf's on both identical length leads > > should all be balanced, and cancel out. In any case, pure copper-copper > > connections generate the lowest thermal emf. I will agree that > manhandling > > 16 gauge solid wire can be very inconvenient. > > > > The path I have taken recently is to order Nakamichi gold over copper > > stereo banana plugs for my connections ( I deal in low voltage work > > exclusively), as well as gold plated spade lugs to go under the banana > > screw-downs. ( My budget ruled out gold over beryllium copper Pomona > brand > > spades. ) I will then experiment between the two connector types. As for > > connections, it seems to me the best course is just to screw down the > > banana plugs, or in the case of the spades, just crimp. I'll wing it on > the > > crimping, and see if simple tools can perform adequately. I would avoid > > solder, since how can one form identical topological spots of solder on > > every connection, deposited at the exact same place on each connection, > and > > ensure the exact same weight of solder, to the microgram, on each > > connection. Since my modus operandi is to aim for balanced emfs, I think > > solder is out. Of course, if one wishes to risk one's health, and that of > > the family, one can track down a source of cadmium, and mix up a lot of > low > > thermal emf solder (cadmium solder is banned in the US, and EU)...not for > > me. > > > > Don't worry, I won't be using tin plated wire, I managed to pick up a > spool > > of silver plated 16 gauge wire for my investigations. Am I on some other > > planet for choosing the balanced thermal emf approach for precision test > > lead applications, as opposed to going to great lengths to eliminate > every > > trace of emf? > > > > Please forgive me if these emf discussions have been between engineers > > working outdoors in the oil fields of North Dakota, or out on a North Sea > > drilling platform. In that case, go for all the super low emf techniques > > you can lay your hands on. > > > > Good Luck, > > Stan > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
TM
Todd Micallef
Wed, Aug 20, 2014 12:50 AM

Stan,

I have seen this Fluke book mentioned occasionally. I am not sure if has
the tutorials that you seek. It is advertised as specific to dc/ lf ac.

http://www.amazon.com/Calibration-Philosophy-Practice-Fluke-Corporation/dp/0963865005

Todd

On Tue, Aug 19, 2014 at 7:35 PM, Stan Katz stan.katz.hk@gmail.com wrote:

You are on the money that guarding/shielding are sadly neglected as
subjects, in their own right. Posts in this group consist primarily of
reports of guard leakage in member instruments, and repair details.
Discussion of proper lash ups of cal. instruments to home built devices, or
even work prototypes, would be very welcome by me. I have read Analog
Devices AN-347, but a tutorial with problems that the student must complete
the lash ups for, would give me a better idea if I've grasped the subject
matter. There are two text books mentioned at the end of AN-347, one by
Morrison, the other by Ott. Any suggestion as to which to buy? The latest
texts by these authors are pricey, so that's why I'd like to pick just one.

Stan, I have seen this Fluke book mentioned occasionally. I am not sure if has the tutorials that you seek. It is advertised as specific to dc/ lf ac. http://www.amazon.com/Calibration-Philosophy-Practice-Fluke-Corporation/dp/0963865005 Todd On Tue, Aug 19, 2014 at 7:35 PM, Stan Katz <stan.katz.hk@gmail.com> wrote: > You are on the money that guarding/shielding are sadly neglected as > subjects, in their own right. Posts in this group consist primarily of > reports of guard leakage in member instruments, and repair details. > Discussion of proper lash ups of cal. instruments to home built devices, or > even work prototypes, would be very welcome by me. I have read Analog > Devices AN-347, but a tutorial with problems that the student must complete > the lash ups for, would give me a better idea if I've grasped the subject > matter. There are two text books mentioned at the end of AN-347, one by > Morrison, the other by Ott. Any suggestion as to which to buy? The latest > texts by these authors are pricey, so that's why I'd like to pick just one. > > > >
SK
Stan Katz
Wed, Aug 20, 2014 1:03 AM

If you reserve those pure copper bananas strictly for infrequent cal. of
something like an HP3458, or other transfer standards in your lab,  they
may be good for some years.

I envision the "beer nuts"  to be a rather relaxed group of individuals,
who are perfectly satisfied to know the alcohol content of their favorite
brew to no better than +/- 60ppm ;-)

Why +/- 60ppm?  A selfish reason. I plan on bringing home a beer-nut-NIST
volt for my Fluke 731B using one of these standards
http://www.ebay.com/itm/10-VOLT-DC-Precision-Voltage-Reference-Standard-Nulled-to-Fluke-732A-732B-/261499015291?pt=LH_DefaultDomain_0&hash=item3ce28e507b

They're only good to 6ppm according to the seller. Rule of thumb is primary
must be ten times the accuracy of secondary, that leaves me with an
uncertainty of +/-60ppm....does seem a bit much....oh well, if necessary,
I'm willing to be the only beer-nuts member.

On Tue, Aug 19, 2014 at 4:12 PM, Todd Micallef tmicallef@gmail.com wrote:

Stan,

I recently picked up some of these ...

http://www.douglasconnection.com/Furez-TSTWP30NP-Bare-Copper-Banana-Plug-Connectors-Pair-FZTSTWP30NP.htm

They are a little pricey and are made for 12ga wire. I think they are OK
for semi-permanent use. A lot of use will probably scratch the soft metal.
I plan on trying them with my Keithley 181 plugged into a low thermal
scanner.
The original Keithley cable will be difficult to terminate. I may have to
go with crimped spade lugs.

Also, I am all for joining a beer nuts group.

Todd

On Tue, Aug 19, 2014 at 2:07 AM, Stan Katz stan.katz.hk@gmail.com wrote:

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my instrumentation
pecking order.  I have all the necessary gear to calibrate these DMMs
according to ancient HP documentation. At the top of my cal. chain is the
731B, called out in HP3456 original documentation.  This hierarchy places
me in the near-beer, or junior member status of the group. I look forward
to being educated, and/or corrected on my understanding of the use of

test

leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use as
thick a solid lead of pure copper wire as you can find, and insert the

wire

into the drill hole on the banana terminal. If the copper is pure, and

has

been properly cleaned, the thermal emf's on both identical length leads
should all be balanced, and cancel out. In any case, pure copper-copper
connections generate the lowest thermal emf. I  will agree that

manhandling

16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over copper
stereo banana plugs for my connections ( I deal in low voltage work
exclusively), as well as gold plated spade lugs to go under the banana
screw-downs. ( My budget ruled out gold over beryllium copper Pomona

brand

spades. )  I will then experiment between the two connector types. As for
connections, it seems to me the best course is just to screw down the
banana plugs, or in the case of the spades, just crimp. I'll wing it on

the

crimping, and see if simple tools can perform adequately. I would avoid
solder, since how can one form identical topological spots of solder on
every connection, deposited at the exact same place on each connection,

and

ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I think
solder is out. Of course, if one wishes to risk one's health, and that of
the family, one can track down a source of cadmium, and mix up a lot of

low

thermal emf solder (cadmium solder is banned in the US, and EU)...not for
me.

Don't worry, I won't be using tin plated wire, I managed to pick up a

spool

of silver plated 16 gauge wire for my investigations.  Am I on some other
planet for choosing the balanced thermal emf approach for precision test
lead applications, as opposed to going to great lengths to eliminate

every

trace of emf?

Please forgive me if these emf discussions have been between engineers
working outdoors in the oil fields of North Dakota, or out on a North Sea
drilling platform. In that case, go for all the super low emf techniques
you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

If you reserve those pure copper bananas strictly for infrequent cal. of something like an HP3458, or other transfer standards in your lab, they may be good for some years. I envision the "beer nuts" to be a rather relaxed group of individuals, who are perfectly satisfied to know the alcohol content of their favorite brew to no better than +/- 60ppm ;-) Why +/- 60ppm? A selfish reason. I plan on bringing home a beer-nut-NIST volt for my Fluke 731B using one of these standards http://www.ebay.com/itm/10-VOLT-DC-Precision-Voltage-Reference-Standard-Nulled-to-Fluke-732A-732B-/261499015291?pt=LH_DefaultDomain_0&hash=item3ce28e507b They're only good to 6ppm according to the seller. Rule of thumb is primary must be ten times the accuracy of secondary, that leaves me with an uncertainty of +/-60ppm....does seem a bit much....oh well, if necessary, I'm willing to be the only beer-nuts member. On Tue, Aug 19, 2014 at 4:12 PM, Todd Micallef <tmicallef@gmail.com> wrote: > Stan, > > I recently picked up some of these ... > > http://www.douglasconnection.com/Furez-TSTWP30NP-Bare-Copper-Banana-Plug-Connectors-Pair-FZTSTWP30NP.htm > > They are a little pricey and are made for 12ga wire. I think they are OK > for semi-permanent use. A lot of use will probably scratch the soft metal. > I plan on trying them with my Keithley 181 plugged into a low thermal > scanner. > The original Keithley cable will be difficult to terminate. I may have to > go with crimped spade lugs. > > Also, I am all for joining a beer nuts group. > > Todd > > > On Tue, Aug 19, 2014 at 2:07 AM, Stan Katz <stan.katz.hk@gmail.com> wrote: > > > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > > HP3458. I do own HP3456 DMMs that are at the top of my instrumentation > > pecking order. I have all the necessary gear to calibrate these DMMs > > according to ancient HP documentation. At the top of my cal. chain is the > > 731B, called out in HP3456 original documentation. This hierarchy places > > me in the near-beer, or junior member status of the group. I look forward > > to being educated, and/or corrected on my understanding of the use of > test > > leads with precision instrumentation. > > > > I don't find much ancient HP documentation on test leads. The only > > recommendations in the era of the 3456 back to the 1960's is to use as > > thick a solid lead of pure copper wire as you can find, and insert the > wire > > into the drill hole on the banana terminal. If the copper is pure, and > has > > been properly cleaned, the thermal emf's on both identical length leads > > should all be balanced, and cancel out. In any case, pure copper-copper > > connections generate the lowest thermal emf. I will agree that > manhandling > > 16 gauge solid wire can be very inconvenient. > > > > The path I have taken recently is to order Nakamichi gold over copper > > stereo banana plugs for my connections ( I deal in low voltage work > > exclusively), as well as gold plated spade lugs to go under the banana > > screw-downs. ( My budget ruled out gold over beryllium copper Pomona > brand > > spades. ) I will then experiment between the two connector types. As for > > connections, it seems to me the best course is just to screw down the > > banana plugs, or in the case of the spades, just crimp. I'll wing it on > the > > crimping, and see if simple tools can perform adequately. I would avoid > > solder, since how can one form identical topological spots of solder on > > every connection, deposited at the exact same place on each connection, > and > > ensure the exact same weight of solder, to the microgram, on each > > connection. Since my modus operandi is to aim for balanced emfs, I think > > solder is out. Of course, if one wishes to risk one's health, and that of > > the family, one can track down a source of cadmium, and mix up a lot of > low > > thermal emf solder (cadmium solder is banned in the US, and EU)...not for > > me. > > > > Don't worry, I won't be using tin plated wire, I managed to pick up a > spool > > of silver plated 16 gauge wire for my investigations. Am I on some other > > planet for choosing the balanced thermal emf approach for precision test > > lead applications, as opposed to going to great lengths to eliminate > every > > trace of emf? > > > > Please forgive me if these emf discussions have been between engineers > > working outdoors in the oil fields of North Dakota, or out on a North Sea > > drilling platform. In that case, go for all the super low emf techniques > > you can lay your hands on. > > > > Good Luck, > > Stan > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
LM
Laurence Motteram
Wed, Aug 20, 2014 1:25 AM

Keithley offer a Low Level Measurements Handbook that may be of interest.  Go to www.keithley.com, Knowledge Centre.  The digital book is a free download (requires registration).  You may be able to get a paperback copy from your local Keithley rep just for the asking.

Regards,

Laurence Motteram

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Stan Katz
Sent: Wednesday, 20 August 2014 9:35 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

You are on the money that guarding/shielding are sadly neglected as subjects, in their own right. Posts in this group consist primarily of reports of guard leakage in member instruments, and repair details.
Discussion of proper lash ups of cal. instruments to home built devices, or even work prototypes, would be very welcome by me. I have read Analog Devices AN-347, but a tutorial with problems that the student must complete the lash ups for, would give me a better idea if I've grasped the subject matter. There are two text books mentioned at the end of AN-347, one by Morrison, the other by Ott. Any suggestion as to which to buy? The latest texts by these authors are pricey, so that's why I'd like to pick just one.

On Tue, Aug 19, 2014 at 3:37 AM, acbern@gmx.de wrote:

generally I think, and at least for your gear, if you use copper plugs
of some type and even solder the wires, thats perfectly fine.
also for a 3458a, and after proper temp stabilization after connecting
the cables (5min) which you always need, these are fine, have not
measrured any difference to crimped cables. also the type of cable
plating (tin/silver..) does not make much difference.
there were many discussions over time, but all the cadmium solder and
non plated copper type debates are not making much sense down to the
3458a level.
I have not done comparison tests with nanovolt meters, as mine have
fixed cables, so no option to change them without destroying them, but
note that e.g. the 182 nanovolt meter has a mil type plug (no special contacts).
sources of error are also things like where to connect the guard to,
these often make a much bigger difference (I have seen these many
times, causing voltage differences of several uV depending where in
the setup you connected the guard, while otherwise the measurement was
perfectly stable), and I bet while people put a lot of effort into the
emf discussion, the guard issue is more often causing errors in
measurements in real life. and thats just one example of additional causes of errors.
cheers

Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr
Von: "Stan Katz" stan.katz.hk@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my
instrumentation pecking order.  I have all the necessary gear to
calibrate these DMMs according to ancient HP documentation. At the
top of my cal. chain is the 731B, called out in HP3456 original
documentation.  This hierarchy places me in the near-beer, or junior
member status of the group. I look forward to being educated, and/or
corrected on my understanding of the use of

test

leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use
as thick a solid lead of pure copper wire as you can find, and
insert the

wire

into the drill hole on the banana terminal. If the copper is pure,
and

has

been properly cleaned, the thermal emf's on both identical length
leads should all be balanced, and cancel out. In any case, pure
copper-copper connections generate the lowest thermal emf. I  will
agree that

manhandling

16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over
copper stereo banana plugs for my connections ( I deal in low
voltage work exclusively), as well as gold plated spade lugs to go
under the banana screw-downs. ( My budget ruled out gold over
beryllium copper Pomona

brand

spades. )  I will then experiment between the two connector types.
As for connections, it seems to me the best course is just to screw
down the banana plugs, or in the case of the spades, just crimp.
I'll wing it on

the

crimping, and see if simple tools can perform adequately. I would
avoid solder, since how can one form identical topological spots of
solder on every connection, deposited at the exact same place on
each connection,

and

ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I
think solder is out. Of course, if one wishes to risk one's health,
and that of the family, one can track down a source of cadmium, and
mix up a lot of

low

thermal emf solder (cadmium solder is banned in the US, and
EU)...not for me.

Don't worry, I won't be using tin plated wire, I managed to pick up
a

spool

of silver plated 16 gauge wire for my investigations.  Am I on some
other planet for choosing the balanced thermal emf approach for
precision test lead applications, as opposed to going to great
lengths to eliminate

every

trace of emf?

Please forgive me if these emf discussions have been between
engineers working outdoors in the oil fields of North Dakota, or out
on a North Sea drilling platform. In that case, go for all the super
low emf techniques you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Message  protected by MailGuard: e-mail anti-virus, anti-spam and content filtering.
http://www.mailguard.com.au

Keithley offer a Low Level Measurements Handbook that may be of interest. Go to www.keithley.com, Knowledge Centre. The digital book is a free download (requires registration). You may be able to get a paperback copy from your local Keithley rep just for the asking. Regards, Laurence Motteram -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Stan Katz Sent: Wednesday, 20 August 2014 9:35 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? You are on the money that guarding/shielding are sadly neglected as subjects, in their own right. Posts in this group consist primarily of reports of guard leakage in member instruments, and repair details. Discussion of proper lash ups of cal. instruments to home built devices, or even work prototypes, would be very welcome by me. I have read Analog Devices AN-347, but a tutorial with problems that the student must complete the lash ups for, would give me a better idea if I've grasped the subject matter. There are two text books mentioned at the end of AN-347, one by Morrison, the other by Ott. Any suggestion as to which to buy? The latest texts by these authors are pricey, so that's why I'd like to pick just one. On Tue, Aug 19, 2014 at 3:37 AM, <acbern@gmx.de> wrote: > generally I think, and at least for your gear, if you use copper plugs > of some type and even solder the wires, thats perfectly fine. > also for a 3458a, and after proper temp stabilization after connecting > the cables (5min) which you always need, these are fine, have not > measrured any difference to crimped cables. also the type of cable > plating (tin/silver..) does not make much difference. > there were many discussions over time, but all the cadmium solder and > non plated copper type debates are not making much sense down to the > 3458a level. > I have not done comparison tests with nanovolt meters, as mine have > fixed cables, so no option to change them without destroying them, but > note that e.g. the 182 nanovolt meter has a mil type plug (no special contacts). > sources of error are also things like where to connect the guard to, > these often make a much bigger difference (I have seen these many > times, causing voltage differences of several uV depending where in > the setup you connected the guard, while otherwise the measurement was > perfectly stable), and I bet while people put a lot of effort into the > emf discussion, the guard issue is more often causing errors in > measurements in real life. and thats just one example of additional causes of errors. > cheers > > > Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr > > Von: "Stan Katz" <stan.katz.hk@gmail.com> > > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? > > > > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > > HP3458. I do own HP3456 DMMs that are at the top of my > > instrumentation pecking order. I have all the necessary gear to > > calibrate these DMMs according to ancient HP documentation. At the > > top of my cal. chain is the 731B, called out in HP3456 original > > documentation. This hierarchy places me in the near-beer, or junior > > member status of the group. I look forward to being educated, and/or > > corrected on my understanding of the use of > test > > leads with precision instrumentation. > > > > I don't find much ancient HP documentation on test leads. The only > > recommendations in the era of the 3456 back to the 1960's is to use > > as thick a solid lead of pure copper wire as you can find, and > > insert the > wire > > into the drill hole on the banana terminal. If the copper is pure, > > and > has > > been properly cleaned, the thermal emf's on both identical length > > leads should all be balanced, and cancel out. In any case, pure > > copper-copper connections generate the lowest thermal emf. I will > > agree that > manhandling > > 16 gauge solid wire can be very inconvenient. > > > > The path I have taken recently is to order Nakamichi gold over > > copper stereo banana plugs for my connections ( I deal in low > > voltage work exclusively), as well as gold plated spade lugs to go > > under the banana screw-downs. ( My budget ruled out gold over > > beryllium copper Pomona > brand > > spades. ) I will then experiment between the two connector types. > > As for connections, it seems to me the best course is just to screw > > down the banana plugs, or in the case of the spades, just crimp. > > I'll wing it on > the > > crimping, and see if simple tools can perform adequately. I would > > avoid solder, since how can one form identical topological spots of > > solder on every connection, deposited at the exact same place on > > each connection, > and > > ensure the exact same weight of solder, to the microgram, on each > > connection. Since my modus operandi is to aim for balanced emfs, I > > think solder is out. Of course, if one wishes to risk one's health, > > and that of the family, one can track down a source of cadmium, and > > mix up a lot of > low > > thermal emf solder (cadmium solder is banned in the US, and > > EU)...not for me. > > > > Don't worry, I won't be using tin plated wire, I managed to pick up > > a > spool > > of silver plated 16 gauge wire for my investigations. Am I on some > > other planet for choosing the balanced thermal emf approach for > > precision test lead applications, as opposed to going to great > > lengths to eliminate > every > > trace of emf? > > > > Please forgive me if these emf discussions have been between > > engineers working outdoors in the oil fields of North Dakota, or out > > on a North Sea drilling platform. In that case, go for all the super > > low emf techniques you can lay your hands on. > > > > Good Luck, > > Stan > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- Message protected by MailGuard: e-mail anti-virus, anti-spam and content filtering. http://www.mailguard.com.au
JH
Joe Hobart
Wed, Aug 20, 2014 1:25 AM

Stan,

Both the Fluke Calibration: Philosophy in Practice and the Keithley: Low Level
Measurements Handbook have chapters about grounding and guarding.  The Low Level
handbook is available from Keithley as a download:

http://www.keithley.com/knowledgecenter/knowledgecenter_pdf/LowLevMsHandbk.pdf

For the 1978-1980 version of the Fluke book:

https://archive.org/details/Calibration-PhilosophyInPractice

The Kindle and epub versions are not very good; get the 32 MB pdf file.

Joe Hobart

On 8/19/2014 5:50 PM, Todd Micallef wrote:

Stan,

I have seen this Fluke book mentioned occasionally. I am not sure if has
the tutorials that you seek. It is advertised as specific to dc/ lf ac.

http://www.amazon.com/Calibration-Philosophy-Practice-Fluke-Corporation/dp/0963865005

Todd

On Tue, Aug 19, 2014 at 7:35 PM, Stan Katz stan.katz.hk@gmail.com wrote:

You are on the money that guarding/shielding are sadly neglected as
subjects, in their own right. Posts in this group consist primarily of
reports of guard leakage in member instruments, and repair details.
Discussion of proper lash ups of cal. instruments to home built devices, or
even work prototypes, would be very welcome by me. I have read Analog
Devices AN-347, but a tutorial with problems that the student must complete
the lash ups for, would give me a better idea if I've grasped the subject
matter. There are two text books mentioned at the end of AN-347, one by
Morrison, the other by Ott. Any suggestion as to which to buy? The latest
texts by these authors are pricey, so that's why I'd like to pick just one.

Stan, Both the Fluke Calibration: Philosophy in Practice and the Keithley: Low Level Measurements Handbook have chapters about grounding and guarding. The Low Level handbook is available from Keithley as a download: http://www.keithley.com/knowledgecenter/knowledgecenter_pdf/LowLevMsHandbk.pdf For the 1978-1980 version of the Fluke book: https://archive.org/details/Calibration-PhilosophyInPractice The Kindle and epub versions are not very good; get the 32 MB pdf file. Joe Hobart On 8/19/2014 5:50 PM, Todd Micallef wrote: > Stan, > > I have seen this Fluke book mentioned occasionally. I am not sure if has > the tutorials that you seek. It is advertised as specific to dc/ lf ac. > > http://www.amazon.com/Calibration-Philosophy-Practice-Fluke-Corporation/dp/0963865005 > > Todd > > > On Tue, Aug 19, 2014 at 7:35 PM, Stan Katz <stan.katz.hk@gmail.com> wrote: > >> You are on the money that guarding/shielding are sadly neglected as >> subjects, in their own right. Posts in this group consist primarily of >> reports of guard leakage in member instruments, and repair details. >> Discussion of proper lash ups of cal. instruments to home built devices, or >> even work prototypes, would be very welcome by me. I have read Analog >> Devices AN-347, but a tutorial with problems that the student must complete >> the lash ups for, would give me a better idea if I've grasped the subject >> matter. There are two text books mentioned at the end of AN-347, one by >> Morrison, the other by Ott. Any suggestion as to which to buy? The latest >> texts by these authors are pricey, so that's why I'd like to pick just one. >> >> >> >>
SG
Stephen Grady
Wed, Aug 20, 2014 4:31 AM

Fellow Volt-Nuts

If I can put a recommendation in for an Australian publication, NMI
Australia runs courses in Electrical Measurements and publishes a monograph
of the course.

Monograph 6 The Measurement of Electrical Quantities (2009, third edition)

Ilya Budovsky et al, 140 pp ($125, overseas $140)
The monograph describes principles and techniques of the measurement of
electrical quantities (direct and alternating voltage, direct and
alternating current, resistance and high voltage), particularly with the use
of digital multimeters (DMMs) and calibrators. The emphasis is on the
techniques of reliable measurement and this necessarily involves a detailed
discussion of the systematic errors that affect the measurement of
electrical quantities as well as the principles of operation of DMMs and
calibrators. Also discussed is the calibration of DMMs and the calculation
of the uncertainty of calibration. The monograph gives an introduction to
the primary standards of electrical quantities.

There is a good section on guarding in this publication; the different types
and correct/incorrect methods of guarding.

It can be ordered through
http://measurement.gov.au/Publications/Pages/Monographs.aspx there is a link
at the top to download a pdf order form.

If you are looking for a free publication I like the Keithley publication
and I vaguely remember Hewlett Packard (as they were then) in the 1970's or
1980's publishing an application note on this subject (a search through the
HP ap note archive may show which one) or Keysight  may still have it on
their website.

Kind Regards,

Stephen Grady
Sydney, Australia

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Stan Katz
Sent: Wednesday, 20 August 2014 9:35 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

You are on the money that guarding/shielding are sadly neglected as
subjects, in their own right. Posts in this group consist primarily of
reports of guard leakage in member instruments, and repair details.
Discussion of proper lash ups of cal. instruments to home built devices, or
even work prototypes, would be very welcome by me. I have read Analog
Devices AN-347, but a tutorial with problems that the student must complete
the lash ups for, would give me a better idea if I've grasped the subject
matter. There are two text books mentioned at the end of AN-347, one by
Morrison, the other by Ott. Any suggestion as to which to buy? The latest
texts by these authors are pricey, so that's why I'd like to pick just one.

On Tue, Aug 19, 2014 at 3:37 AM, acbern@gmx.de wrote:

generally I think, and at least for your gear, if you use copper plugs
of some type and even solder the wires, thats perfectly fine.
also for a 3458a, and after proper temp stabilization after connecting
the cables (5min) which you always need, these are fine, have not
measrured any difference to crimped cables. also the type of cable
plating (tin/silver..) does not make much difference.
there were many discussions over time, but all the cadmium solder and
non plated copper type debates are not making much sense down to the
3458a level.
I have not done comparison tests with nanovolt meters, as mine have
fixed cables, so no option to change them without destroying them, but
note that e.g. the 182 nanovolt meter has a mil type plug (no special

contacts).

sources of error are also things like where to connect the guard to,
these often make a much bigger difference (I have seen these many
times, causing voltage differences of several uV depending where in
the setup you connected the guard, while otherwise the measurement was
perfectly stable), and I bet while people put a lot of effort into the
emf discussion, the guard issue is more often causing errors in
measurements in real life. and thats just one example of additional causes

of errors.

cheers

Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr
Von: "Stan Katz" stan.katz.hk@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my
instrumentation pecking order.  I have all the necessary gear to
calibrate these DMMs according to ancient HP documentation. At the
top of my cal. chain is the 731B, called out in HP3456 original
documentation.  This hierarchy places me in the near-beer, or junior
member status of the group. I look forward to being educated, and/or
corrected on my understanding of the use of

test

leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use
as thick a solid lead of pure copper wire as you can find, and
insert the

wire

into the drill hole on the banana terminal. If the copper is pure,
and

has

been properly cleaned, the thermal emf's on both identical length
leads should all be balanced, and cancel out. In any case, pure
copper-copper connections generate the lowest thermal emf. I  will
agree that

manhandling

16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over
copper stereo banana plugs for my connections ( I deal in low
voltage work exclusively), as well as gold plated spade lugs to go
under the banana screw-downs. ( My budget ruled out gold over
beryllium copper Pomona

brand

spades. )  I will then experiment between the two connector types.
As for connections, it seems to me the best course is just to screw
down the banana plugs, or in the case of the spades, just crimp.
I'll wing it on

the

crimping, and see if simple tools can perform adequately. I would
avoid solder, since how can one form identical topological spots of
solder on every connection, deposited at the exact same place on
each connection,

and

ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I
think solder is out. Of course, if one wishes to risk one's health,
and that of the family, one can track down a source of cadmium, and
mix up a lot of

low

thermal emf solder (cadmium solder is banned in the US, and
EU)...not for me.

Don't worry, I won't be using tin plated wire, I managed to pick up
a

spool

of silver plated 16 gauge wire for my investigations.  Am I on some
other planet for choosing the balanced thermal emf approach for
precision test lead applications, as opposed to going to great
lengths to eliminate

every

trace of emf?

Please forgive me if these emf discussions have been between
engineers working outdoors in the oil fields of North Dakota, or out
on a North Sea drilling platform. In that case, go for all the super
low emf techniques you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com

Fellow Volt-Nuts If I can put a recommendation in for an Australian publication, NMI Australia runs courses in Electrical Measurements and publishes a monograph of the course. Monograph 6 The Measurement of Electrical Quantities (2009, third edition) Ilya Budovsky et al, 140 pp ($125, overseas $140) The monograph describes principles and techniques of the measurement of electrical quantities (direct and alternating voltage, direct and alternating current, resistance and high voltage), particularly with the use of digital multimeters (DMMs) and calibrators. The emphasis is on the techniques of reliable measurement and this necessarily involves a detailed discussion of the systematic errors that affect the measurement of electrical quantities as well as the principles of operation of DMMs and calibrators. Also discussed is the calibration of DMMs and the calculation of the uncertainty of calibration. The monograph gives an introduction to the primary standards of electrical quantities. There is a good section on guarding in this publication; the different types and correct/incorrect methods of guarding. It can be ordered through http://measurement.gov.au/Publications/Pages/Monographs.aspx there is a link at the top to download a pdf order form. If you are looking for a free publication I like the Keithley publication and I vaguely remember Hewlett Packard (as they were then) in the 1970's or 1980's publishing an application note on this subject (a search through the HP ap note archive may show which one) or Keysight may still have it on their website. Kind Regards, Stephen Grady Sydney, Australia -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Stan Katz Sent: Wednesday, 20 August 2014 9:35 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? You are on the money that guarding/shielding are sadly neglected as subjects, in their own right. Posts in this group consist primarily of reports of guard leakage in member instruments, and repair details. Discussion of proper lash ups of cal. instruments to home built devices, or even work prototypes, would be very welcome by me. I have read Analog Devices AN-347, but a tutorial with problems that the student must complete the lash ups for, would give me a better idea if I've grasped the subject matter. There are two text books mentioned at the end of AN-347, one by Morrison, the other by Ott. Any suggestion as to which to buy? The latest texts by these authors are pricey, so that's why I'd like to pick just one. On Tue, Aug 19, 2014 at 3:37 AM, <acbern@gmx.de> wrote: > generally I think, and at least for your gear, if you use copper plugs > of some type and even solder the wires, thats perfectly fine. > also for a 3458a, and after proper temp stabilization after connecting > the cables (5min) which you always need, these are fine, have not > measrured any difference to crimped cables. also the type of cable > plating (tin/silver..) does not make much difference. > there were many discussions over time, but all the cadmium solder and > non plated copper type debates are not making much sense down to the > 3458a level. > I have not done comparison tests with nanovolt meters, as mine have > fixed cables, so no option to change them without destroying them, but > note that e.g. the 182 nanovolt meter has a mil type plug (no special contacts). > sources of error are also things like where to connect the guard to, > these often make a much bigger difference (I have seen these many > times, causing voltage differences of several uV depending where in > the setup you connected the guard, while otherwise the measurement was > perfectly stable), and I bet while people put a lot of effort into the > emf discussion, the guard issue is more often causing errors in > measurements in real life. and thats just one example of additional causes of errors. > cheers > > > Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr > > Von: "Stan Katz" <stan.katz.hk@gmail.com> > > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? > > > > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > > HP3458. I do own HP3456 DMMs that are at the top of my > > instrumentation pecking order. I have all the necessary gear to > > calibrate these DMMs according to ancient HP documentation. At the > > top of my cal. chain is the 731B, called out in HP3456 original > > documentation. This hierarchy places me in the near-beer, or junior > > member status of the group. I look forward to being educated, and/or > > corrected on my understanding of the use of > test > > leads with precision instrumentation. > > > > I don't find much ancient HP documentation on test leads. The only > > recommendations in the era of the 3456 back to the 1960's is to use > > as thick a solid lead of pure copper wire as you can find, and > > insert the > wire > > into the drill hole on the banana terminal. If the copper is pure, > > and > has > > been properly cleaned, the thermal emf's on both identical length > > leads should all be balanced, and cancel out. In any case, pure > > copper-copper connections generate the lowest thermal emf. I will > > agree that > manhandling > > 16 gauge solid wire can be very inconvenient. > > > > The path I have taken recently is to order Nakamichi gold over > > copper stereo banana plugs for my connections ( I deal in low > > voltage work exclusively), as well as gold plated spade lugs to go > > under the banana screw-downs. ( My budget ruled out gold over > > beryllium copper Pomona > brand > > spades. ) I will then experiment between the two connector types. > > As for connections, it seems to me the best course is just to screw > > down the banana plugs, or in the case of the spades, just crimp. > > I'll wing it on > the > > crimping, and see if simple tools can perform adequately. I would > > avoid solder, since how can one form identical topological spots of > > solder on every connection, deposited at the exact same place on > > each connection, > and > > ensure the exact same weight of solder, to the microgram, on each > > connection. Since my modus operandi is to aim for balanced emfs, I > > think solder is out. Of course, if one wishes to risk one's health, > > and that of the family, one can track down a source of cadmium, and > > mix up a lot of > low > > thermal emf solder (cadmium solder is banned in the US, and > > EU)...not for me. > > > > Don't worry, I won't be using tin plated wire, I managed to pick up > > a > spool > > of silver plated 16 gauge wire for my investigations. Am I on some > > other planet for choosing the balanced thermal emf approach for > > precision test lead applications, as opposed to going to great > > lengths to eliminate > every > > trace of emf? > > > > Please forgive me if these emf discussions have been between > > engineers working outdoors in the oil fields of North Dakota, or out > > on a North Sea drilling platform. In that case, go for all the super > > low emf techniques you can lay your hands on. > > > > Good Luck, > > Stan > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com
P
pa4tim@gmail.com
Wed, Aug 20, 2014 6:23 AM

About the Keithley mentioned by stephen, It is a free download but they have a free paper version too.  I have a picture of it on my site. Scroll down until you see a red book. http://www.pa4tim.nl/?p=2899

Fred, pa4tim

Verzonden met Windows Mail

Van: Stephen Grady
Verzonden: ‎woensdag‎ ‎20‎ ‎augustus‎ ‎2014 ‎06‎:‎31
Aan: volt-nuts

Fellow Volt-Nuts

If I can put a recommendation in for an Australian publication, NMI
Australia runs courses in Electrical Measurements and publishes a monograph
of the course.

Monograph 6 The Measurement of Electrical Quantities (2009, third edition)

Ilya Budovsky et al, 140 pp ($125, overseas $140)
The monograph describes principles and techniques of the measurement of
electrical quantities (direct and alternating voltage, direct and
alternating current, resistance and high voltage), particularly with the use
of digital multimeters (DMMs) and calibrators. The emphasis is on the
techniques of reliable measurement and this necessarily involves a detailed
discussion of the systematic errors that affect the measurement of
electrical quantities as well as the principles of operation of DMMs and
calibrators. Also discussed is the calibration of DMMs and the calculation
of the uncertainty of calibration. The monograph gives an introduction to
the primary standards of electrical quantities.

There is a good section on guarding in this publication; the different types
and correct/incorrect methods of guarding.

It can be ordered through
http://measurement.gov.au/Publications/Pages/Monographs.aspx there is a link
at the top to download a pdf order form.

If you are looking for a free publication I like the Keithley publication
and I vaguely remember Hewlett Packard (as they were then) in the 1970's or
1980's publishing an application note on this subject (a search through the
HP ap note archive may show which one) or Keysight  may still have it on
their website.

Kind Regards,

Stephen Grady
Sydney, Australia

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Stan Katz
Sent: Wednesday, 20 August 2014 9:35 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

You are on the money that guarding/shielding are sadly neglected as
subjects, in their own right. Posts in this group consist primarily of
reports of guard leakage in member instruments, and repair details.
Discussion of proper lash ups of cal. instruments to home built devices, or
even work prototypes, would be very welcome by me. I have read Analog
Devices AN-347, but a tutorial with problems that the student must complete
the lash ups for, would give me a better idea if I've grasped the subject
matter. There are two text books mentioned at the end of AN-347, one by
Morrison, the other by Ott. Any suggestion as to which to buy? The latest
texts by these authors are pricey, so that's why I'd like to pick just one.

On Tue, Aug 19, 2014 at 3:37 AM, acbern@gmx.de wrote:

generally I think, and at least for your gear, if you use copper plugs
of some type and even solder the wires, thats perfectly fine.
also for a 3458a, and after proper temp stabilization after connecting
the cables (5min) which you always need, these are fine, have not
measrured any difference to crimped cables. also the type of cable
plating (tin/silver..) does not make much difference.
there were many discussions over time, but all the cadmium solder and
non plated copper type debates are not making much sense down to the
3458a level.
I have not done comparison tests with nanovolt meters, as mine have
fixed cables, so no option to change them without destroying them, but
note that e.g. the 182 nanovolt meter has a mil type plug (no special

contacts).

sources of error are also things like where to connect the guard to,
these often make a much bigger difference (I have seen these many
times, causing voltage differences of several uV depending where in
the setup you connected the guard, while otherwise the measurement was
perfectly stable), and I bet while people put a lot of effort into the
emf discussion, the guard issue is more often causing errors in
measurements in real life. and thats just one example of additional causes

of errors.

cheers

Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr
Von: "Stan Katz" stan.katz.hk@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff?

I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an
HP3458. I do own HP3456 DMMs that are at the top of my
instrumentation pecking order.  I have all the necessary gear to
calibrate these DMMs according to ancient HP documentation. At the
top of my cal. chain is the 731B, called out in HP3456 original
documentation.  This hierarchy places me in the near-beer, or junior
member status of the group. I look forward to being educated, and/or
corrected on my understanding of the use of

test

leads with precision instrumentation.

I don't find much ancient HP documentation on test leads. The only
recommendations in the era of the 3456 back to the 1960's is to use
as thick a solid lead of pure copper wire as you can find, and
insert the

wire

into the drill hole on the banana terminal. If the copper is pure,
and

has

been properly cleaned, the thermal emf's on both identical length
leads should all be balanced, and cancel out. In any case, pure
copper-copper connections generate the lowest thermal emf. I  will
agree that

manhandling

16 gauge solid wire can be very inconvenient.

The path I have taken recently is to order Nakamichi gold over
copper stereo banana plugs for my connections ( I deal in low
voltage work exclusively), as well as gold plated spade lugs to go
under the banana screw-downs. ( My budget ruled out gold over
beryllium copper Pomona

brand

spades. )  I will then experiment between the two connector types.
As for connections, it seems to me the best course is just to screw
down the banana plugs, or in the case of the spades, just crimp.
I'll wing it on

the

crimping, and see if simple tools can perform adequately. I would
avoid solder, since how can one form identical topological spots of
solder on every connection, deposited at the exact same place on
each connection,

and

ensure the exact same weight of solder, to the microgram, on each
connection. Since my modus operandi is to aim for balanced emfs, I
think solder is out. Of course, if one wishes to risk one's health,
and that of the family, one can track down a source of cadmium, and
mix up a lot of

low

thermal emf solder (cadmium solder is banned in the US, and
EU)...not for me.

Don't worry, I won't be using tin plated wire, I managed to pick up
a

spool

of silver plated 16 gauge wire for my investigations.  Am I on some
other planet for choosing the balanced thermal emf approach for
precision test lead applications, as opposed to going to great
lengths to eliminate

every

trace of emf?

Please forgive me if these emf discussions have been between
engineers working outdoors in the oil fields of North Dakota, or out
on a North Sea drilling platform. In that case, go for all the super
low emf techniques you can lay your hands on.

Good Luck,
Stan


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

About the Keithley mentioned by stephen, It is a free download but they have a free paper version too. I have a picture of it on my site. Scroll down until you see a red book. http://www.pa4tim.nl/?p=2899 Fred, pa4tim Verzonden met Windows Mail Van: Stephen Grady Verzonden: ‎woensdag‎ ‎20‎ ‎augustus‎ ‎2014 ‎06‎:‎31 Aan: volt-nuts Fellow Volt-Nuts If I can put a recommendation in for an Australian publication, NMI Australia runs courses in Electrical Measurements and publishes a monograph of the course. Monograph 6 The Measurement of Electrical Quantities (2009, third edition) Ilya Budovsky et al, 140 pp ($125, overseas $140) The monograph describes principles and techniques of the measurement of electrical quantities (direct and alternating voltage, direct and alternating current, resistance and high voltage), particularly with the use of digital multimeters (DMMs) and calibrators. The emphasis is on the techniques of reliable measurement and this necessarily involves a detailed discussion of the systematic errors that affect the measurement of electrical quantities as well as the principles of operation of DMMs and calibrators. Also discussed is the calibration of DMMs and the calculation of the uncertainty of calibration. The monograph gives an introduction to the primary standards of electrical quantities. There is a good section on guarding in this publication; the different types and correct/incorrect methods of guarding. It can be ordered through http://measurement.gov.au/Publications/Pages/Monographs.aspx there is a link at the top to download a pdf order form. If you are looking for a free publication I like the Keithley publication and I vaguely remember Hewlett Packard (as they were then) in the 1970's or 1980's publishing an application note on this subject (a search through the HP ap note archive may show which one) or Keysight may still have it on their website. Kind Regards, Stephen Grady Sydney, Australia -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Stan Katz Sent: Wednesday, 20 August 2014 9:35 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? You are on the money that guarding/shielding are sadly neglected as subjects, in their own right. Posts in this group consist primarily of reports of guard leakage in member instruments, and repair details. Discussion of proper lash ups of cal. instruments to home built devices, or even work prototypes, would be very welcome by me. I have read Analog Devices AN-347, but a tutorial with problems that the student must complete the lash ups for, would give me a better idea if I've grasped the subject matter. There are two text books mentioned at the end of AN-347, one by Morrison, the other by Ott. Any suggestion as to which to buy? The latest texts by these authors are pricey, so that's why I'd like to pick just one. On Tue, Aug 19, 2014 at 3:37 AM, <acbern@gmx.de> wrote: > generally I think, and at least for your gear, if you use copper plugs > of some type and even solder the wires, thats perfectly fine. > also for a 3458a, and after proper temp stabilization after connecting > the cables (5min) which you always need, these are fine, have not > measrured any difference to crimped cables. also the type of cable > plating (tin/silver..) does not make much difference. > there were many discussions over time, but all the cadmium solder and > non plated copper type debates are not making much sense down to the > 3458a level. > I have not done comparison tests with nanovolt meters, as mine have > fixed cables, so no option to change them without destroying them, but > note that e.g. the 182 nanovolt meter has a mil type plug (no special contacts). > sources of error are also things like where to connect the guard to, > these often make a much bigger difference (I have seen these many > times, causing voltage differences of several uV depending where in > the setup you connected the guard, while otherwise the measurement was > perfectly stable), and I bet while people put a lot of effort into the > emf discussion, the guard issue is more often causing errors in > measurements in real life. and thats just one example of additional causes of errors. > cheers > > > Gesendet: Dienstag, 19. August 2014 um 06:07 Uhr > > Von: "Stan Katz" <stan.katz.hk@gmail.com> > > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > Betreff: [volt-nuts] What's All This Low Thermal EMF Test Lead Stuff? > > > > I'm a self described volt-nut-near-beer. I don't own a 732A/B, or an > > HP3458. I do own HP3456 DMMs that are at the top of my > > instrumentation pecking order. I have all the necessary gear to > > calibrate these DMMs according to ancient HP documentation. At the > > top of my cal. chain is the 731B, called out in HP3456 original > > documentation. This hierarchy places me in the near-beer, or junior > > member status of the group. I look forward to being educated, and/or > > corrected on my understanding of the use of > test > > leads with precision instrumentation. > > > > I don't find much ancient HP documentation on test leads. The only > > recommendations in the era of the 3456 back to the 1960's is to use > > as thick a solid lead of pure copper wire as you can find, and > > insert the > wire > > into the drill hole on the banana terminal. If the copper is pure, > > and > has > > been properly cleaned, the thermal emf's on both identical length > > leads should all be balanced, and cancel out. In any case, pure > > copper-copper connections generate the lowest thermal emf. I will > > agree that > manhandling > > 16 gauge solid wire can be very inconvenient. > > > > The path I have taken recently is to order Nakamichi gold over > > copper stereo banana plugs for my connections ( I deal in low > > voltage work exclusively), as well as gold plated spade lugs to go > > under the banana screw-downs. ( My budget ruled out gold over > > beryllium copper Pomona > brand > > spades. ) I will then experiment between the two connector types. > > As for connections, it seems to me the best course is just to screw > > down the banana plugs, or in the case of the spades, just crimp. > > I'll wing it on > the > > crimping, and see if simple tools can perform adequately. I would > > avoid solder, since how can one form identical topological spots of > > solder on every connection, deposited at the exact same place on > > each connection, > and > > ensure the exact same weight of solder, to the microgram, on each > > connection. Since my modus operandi is to aim for balanced emfs, I > > think solder is out. Of course, if one wishes to risk one's health, > > and that of the family, one can track down a source of cadmium, and > > mix up a lot of > low > > thermal emf solder (cadmium solder is banned in the US, and > > EU)...not for me. > > > > Don't worry, I won't be using tin plated wire, I managed to pick up > > a > spool > > of silver plated 16 gauge wire for my investigations. Am I on some > > other planet for choosing the balanced thermal emf approach for > > precision test lead applications, as opposed to going to great > > lengths to eliminate > every > > trace of emf? > > > > Please forgive me if these emf discussions have been between > > engineers working outdoors in the oil fields of North Dakota, or out > > on a North Sea drilling platform. In that case, go for all the super > > low emf techniques you can lay your hands on. > > > > Good Luck, > > Stan > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.