As part of my checking over of the behaviour of the 7081 I ran "mode,testk" and "nines=8" overnight, the result is a swing larger than I would expect!
max 166.00455
min 166.00040
this means a 25 ppm shift
The circuit is meant to generate a constant current, so if that constant current can vary by 25ppm over a 20-23 degree house temperature then it is only accurate to 25ppm for resistance measurement except when used as a transfer standard.
anyone else find a similar shift?
M K
How long had the meter been powered on for when you started the test? Think days rather than hours to stabilise.
D.
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k
Sent: 20 January 2013 09:06
To: volt-nuts@febo.com
Subject: [volt-nuts] 7081 resistance accuracy issue
Hi everyone,
As part of my checking over of the behaviour of the 7081 I ran "mode,testk" and "nines=8" overnight, the result is a swing larger than I would expect!
max 166.00455
min 166.00040
this means a 25 ppm shift
The circuit is meant to generate a constant current, so if that constant current can vary by 25ppm over a 20-23 degree house temperature then it is only accurate to 25ppm for resistance measurement except when used as a transfer standard.
anyone else find a similar shift?
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi David,
Yes, the meter has been on for days already. It had just done a long run into a shorting plug that shows much better stability, I suspect a tempco in the opto-isolators that transfer the reference volts, but will carefully consider all possible sources first.
The other option is the input bias current that Mickle found.
Regards,
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 20 Jan 2013 11:13:00 +0000
Subject: Re: [volt-nuts] 7081 resistance accuracy issue
How long had the meter been powered on for when you started the test? Think days rather than hours to stabilise.
D.
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k
Sent: 20 January 2013 09:06
To: volt-nuts@febo.com
Subject: [volt-nuts] 7081 resistance accuracy issue
Hi everyone,
As part of my checking over of the behaviour of the 7081 I ran "mode,testk" and "nines=8" overnight, the result is a swing larger than I would expect!
max 166.00455
min 166.00040
this means a 25 ppm shift
The circuit is meant to generate a constant current, so if that constant current can vary by 25ppm over a 20-23 degree house temperature then it is only accurate to 25ppm for resistance measurement except when used as a transfer standard.
anyone else find a similar shift?
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi, M K!
It is almost impossible to check any of the Ohms range performance via TESTK
mode. While in this mode a constant current is applied to the ADC with a low
input resistance equal to 160 kOhm (R201). R201 paired with R202 is a Vishay
VHA51x custom metalfoil resistors set. It have a very low TCR tracking error
(needed for ADC slope stability), but a large absolute TCR value, up to 5 ppm/C.
Regards,
Mickle T.
Hi Mickle,
That is good to know, I know it is a positive tempco I have, I will just have to try a stable resistor and see what it gives me, but there is more checking to do first.
Regards,
M K
Date: Sun, 20 Jan 2013 17:02:04 +0400
From: timka2k@yandex.ru
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 7081 resistance accuracy issue
Hi, M K!
It is almost impossible to check any of the Ohms range performance via TESTK
mode. While in this mode a constant current is applied to the ADC with a low
input resistance equal to 160 kOhm (R201). R201 paired with R202 is a Vishay
VHA51x custom metalfoil resistors set. It have a very low TCR tracking error
(needed for ADC slope stability), but a large absolute TCR value, up to 5 ppm/C.
Regards,
Mickle T.