HP/Agilet counter 53151A trigger problems

BN
Bernd Neubig
Fri, Aug 24, 2012 1:34 PM

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in my lab.
Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three input
levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input level
rather selectively, i.e. mostly at 100 MHz, some also at 150 MHz, while the
are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com

Hi all, I have several HP/Agilent frequency counters 53151A (and 53152A) in my lab. Several of them meanwhile show trigger problems. I have tested the A and B channel between 5 MHz and 200 MHZ at three input levels of -10 dBm, 0 dBm and +10 dBm. Those with trigger problems show erroneous numbers at -10 dBm input level rather selectively, i.e. mostly at 100 MHz, some also at 150 MHz, while the are o.k. Has anyone on this list experienced similar defects and can give a recommendation how to fix/repair them? Best regards Bernd DK1AG www.axtal.com
AB
Azelio Boriani
Fri, Aug 24, 2012 3:28 PM

From the datasheet:

A input 10Hz..125MHz min level 25mVrms (-19dBm)
B input min 50MHz, -20dBm
the problem is on A input only? Maybe 150MHz for the A input, being beyond
the official specifications, requires more signal on one counter than the
other. In the past they all were fine?
I have read your paper on the correlation between the real and predicted
aging of crystal oscillators. I'm trying to find a suitable model for the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in my lab.
Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three input
levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input level
rather selectively, i.e. mostly at 100 MHz, some also at 150 MHz, while the
are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

>From the datasheet: A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm the problem is on A input only? Maybe 150MHz for the A input, being beyond the official specifications, requires more signal on one counter than the other. In the past they all were fine? I have read your paper on the correlation between the real and predicted aging of crystal oscillators. I'm trying to find a suitable model for the Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter. On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> wrote: > Hi all, > > I have several HP/Agilent frequency counters 53151A (and 53152A) in my lab. > Several of them meanwhile show trigger problems. > > I have tested the A and B channel between 5 MHz and 200 MHZ at three input > levels of -10 dBm, 0 dBm and +10 dBm. > > Those with trigger problems show erroneous numbers at -10 dBm input level > rather selectively, i.e. mostly at 100 MHz, some also at 150 MHz, while the > are o.k. > > Has anyone on this list experienced similar defects and can give a > recommendation how to fix/repair them? > > > > Best regards > > > > Bernd > > DK1AG > > www.axtal.com > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BN
Bernd Neubig
Mon, Aug 27, 2012 5:01 PM

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, <30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in the range
between 100 MHz and <200 MHz, observed in automatic trigger mode. In manual
trigger mode the counter does not trigger at all in that freq range.

I cannot tell you an aging model for a particular OCXO, and I doubt if
anyone can ;) Aging predictions can only be made on individual OCXO, unit by
unit, and the prediction is only valid over a rather limited time. Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:

A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm
the problem is on A input only? Maybe 150MHz for the A input, being beyond
the official specifications, requires more signal on one counter than the
other. In the past they all were fine?
I have read your paper on the correlation between the real and predicted
aging of crystal oscillators. I'm trying to find a suitable model for the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in my

lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three
input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input
level rather selectively, i.e. mostly at 100 MHz, some also at 150
MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Azelio, Sorry, the counters are 53131A, not 53151A. The A and B inputs are specified from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, <30 mV (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. The problem can be at either input alone. The problem is only in the range between 100 MHz and <200 MHz, observed in automatic trigger mode. In manual trigger mode the counter does not trigger at all in that freq range. I cannot tell you an aging model for a particular OCXO, and I doubt if anyone can ;) Aging predictions can only be made on individual OCXO, unit by unit, and the prediction is only valid over a rather limited time. Therefore you must continuously monitor and update your prediction. Best regards Bernd DK1AG -----Ursprüngliche Nachricht----- Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im Auftrag von Azelio Boriani Gesendet: Freitag, 24. August 2012 17:28 An: Discussion of precise time and frequency measurement Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems >From the datasheet: A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm the problem is on A input only? Maybe 150MHz for the A input, being beyond the official specifications, requires more signal on one counter than the other. In the past they all were fine? I have read your paper on the correlation between the real and predicted aging of crystal oscillators. I'm trying to find a suitable model for the Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter. On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> wrote: > Hi all, > > I have several HP/Agilent frequency counters 53151A (and 53152A) in my lab. > Several of them meanwhile show trigger problems. > > I have tested the A and B channel between 5 MHz and 200 MHZ at three > input levels of -10 dBm, 0 dBm and +10 dBm. > > Those with trigger problems show erroneous numbers at -10 dBm input > level rather selectively, i.e. mostly at 100 MHz, some also at 150 > MHz, while the are o.k. > > Has anyone on this list experienced similar defects and can give a > recommendation how to fix/repair them? > > > > Best regards > > > > Bernd > > DK1AG > > www.axtal.com > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
AB
Azelio Boriani
Mon, Aug 27, 2012 7:19 PM

30mV are -13dBm so maybe it is why the -20dBm is not a suitable level for
that input. I have the 53132A and the 53181A so I can test (with the R&S
SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a model that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the parameters) will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, <30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in the range
between 100 MHz and <200 MHz, observed in automatic trigger mode. In manual
trigger mode the counter does not trigger at all in that freq range.

I cannot tell you an aging model for a particular OCXO, and I doubt if
anyone can ;) Aging predictions can only be made on individual OCXO, unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm
the problem is on A input only? Maybe 150MHz for the A input, being beyond
the official specifications, requires more signal on one counter than the
other. In the past they all were fine?
I have read your paper on the correlation between the real and predicted
aging of crystal oscillators. I'm trying to find a suitable model for the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in my

lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three
input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input
level rather selectively, i.e. mostly at 100 MHz, some also at 150
MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

30mV are -13dBm so maybe it is why the -20dBm is not a suitable level for that input. I have the 53132A and the 53181A so I can test (with the R&S SML01). Is this problem new? Never tested before? About the OCXO aging: OK, I'm interested in the development of a model that can be general enough so that it can be "steered", adjusting its parameters, during the OCXO life. The data (to adjust the parameters) will be gathered by the continuous monitoring with the GPS. On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> wrote: > Hi Azelio, > > Sorry, the counters are 53131A, not 53151A. The A and B inputs are > specified > from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, <30 mV > (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. > The problem can be at either input alone. The problem is only in the range > between 100 MHz and <200 MHz, observed in automatic trigger mode. In manual > trigger mode the counter does not trigger at all in that freq range. > > I cannot tell you an aging model for a particular OCXO, and I doubt if > anyone can ;) Aging predictions can only be made on individual OCXO, unit > by > unit, and the prediction is only valid over a rather limited time. > Therefore > you must continuously monitor and update your prediction. > > Best regards > > Bernd > DK1AG > > -----Ursprüngliche Nachricht----- > Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im > Auftrag von Azelio Boriani > Gesendet: Freitag, 24. August 2012 17:28 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems > > From the datasheet: > A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm > the problem is on A input only? Maybe 150MHz for the A input, being beyond > the official specifications, requires more signal on one counter than the > other. In the past they all were fine? > I have read your paper on the correlation between the real and predicted > aging of crystal oscillators. I'm trying to find a suitable model for the > Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter. > > On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> wrote: > > > Hi all, > > > > I have several HP/Agilent frequency counters 53151A (and 53152A) in my > lab. > > Several of them meanwhile show trigger problems. > > > > I have tested the A and B channel between 5 MHz and 200 MHZ at three > > input levels of -10 dBm, 0 dBm and +10 dBm. > > > > Those with trigger problems show erroneous numbers at -10 dBm input > > level rather selectively, i.e. mostly at 100 MHz, some also at 150 > > MHz, while the are o.k. > > > > Has anyone on this list experienced similar defects and can give a > > recommendation how to fix/repair them? > > > > > > > > Best regards > > > > > > > > Bernd > > > > DK1AG > > > > www.axtal.com > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
DL
Don Latham
Mon, Aug 27, 2012 7:31 PM

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable level
for
that input. I have the 53132A and the 53181A so I can test (with the R&S
SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a model
that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the parameters)
will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de
wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz,
<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in the
range
between 100 MHz and <200 MHz, observed in automatic trigger mode. In
manual
trigger mode the counter does not trigger at all in that freq range.

I cannot tell you an aging model for a particular OCXO, and I doubt if
anyone can ;) Aging predictions can only be made on individual OCXO,
unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz,
-20dBm
the problem is on A input only? Maybe 150MHz for the A input, being
beyond
the official specifications, requires more signal on one counter than
the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable model for
the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman
filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de
wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three
input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input
level rather selectively, i.e. mostly at 100 MHz, some also at 150
MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

Isn't crystal aging marked by random jumps? perhaps lattice rearrangement? Really hard if not impossible to model... Don Azelio Boriani > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable level > for > that input. I have the 53132A and the 53181A so I can test (with the R&S > SML01). Is this problem new? Never tested before? > > About the OCXO aging: OK, I'm interested in the development of a model > that > can be general enough so that it can be "steered", adjusting its > parameters, during the OCXO life. The data (to adjust the parameters) > will > be gathered by the continuous monitoring with the GPS. > > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> > wrote: > >> Hi Azelio, >> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs are >> specified >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, >> <30 mV >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. >> The problem can be at either input alone. The problem is only in the >> range >> between 100 MHz and <200 MHz, observed in automatic trigger mode. In >> manual >> trigger mode the counter does not trigger at all in that freq range. >> >> I cannot tell you an aging model for a particular OCXO, and I doubt if >> anyone can ;) Aging predictions can only be made on individual OCXO, >> unit >> by >> unit, and the prediction is only valid over a rather limited time. >> Therefore >> you must continuously monitor and update your prediction. >> >> Best regards >> >> Bernd >> DK1AG >> >> -----Ursprüngliche Nachricht----- >> Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im >> Auftrag von Azelio Boriani >> Gesendet: Freitag, 24. August 2012 17:28 >> An: Discussion of precise time and frequency measurement >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems >> >> From the datasheet: >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, >> -20dBm >> the problem is on A input only? Maybe 150MHz for the A input, being >> beyond >> the official specifications, requires more signal on one counter than >> the >> other. In the past they all were fine? >> I have read your paper on the correlation between the real and >> predicted >> aging of crystal oscillators. I'm trying to find a suitable model for >> the >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman >> filter. >> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> >> wrote: >> >> > Hi all, >> > >> > I have several HP/Agilent frequency counters 53151A (and 53152A) in >> my >> lab. >> > Several of them meanwhile show trigger problems. >> > >> > I have tested the A and B channel between 5 MHz and 200 MHZ at three >> > input levels of -10 dBm, 0 dBm and +10 dBm. >> > >> > Those with trigger problems show erroneous numbers at -10 dBm input >> > level rather selectively, i.e. mostly at 100 MHz, some also at 150 >> > MHz, while the are o.k. >> > >> > Has anyone on this list experienced similar defects and can give a >> > recommendation how to fix/repair them? >> > >> > >> > >> > Best regards >> > >> > >> > >> > Bernd >> > >> > DK1AG >> > >> > www.axtal.com >> > >> > _______________________________________________ >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> > and follow the instructions there. >> > >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." R. Bacon "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
AB
Azelio Boriani
Mon, Aug 27, 2012 7:41 PM

So it can't be done... the possibility to discipline an OCXO taking the
action from a suitable model should help in speeding up the adjustments and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable level
for
that input. I have the 53132A and the 53181A so I can test (with the R&S
SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a model
that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the parameters)
will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de
wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz,
<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in the
range
between 100 MHz and <200 MHz, observed in automatic trigger mode. In
manual
trigger mode the counter does not trigger at all in that freq range.

I cannot tell you an aging model for a particular OCXO, and I doubt if
anyone can ;) Aging predictions can only be made on individual OCXO,
unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz,
-20dBm
the problem is on A input only? Maybe 150MHz for the A input, being
beyond
the official specifications, requires more signal on one counter than
the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable model for
the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman
filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de
wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three
input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input
level rather selectively, i.e. mostly at 100 MHz, some also at 150
MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

So it can't be done... the possibility to discipline an OCXO taking the action from a suitable model should help in speeding up the adjustments and avoiding the humps as seen in the Allan deviation plot. On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: > Isn't crystal aging marked by random jumps? perhaps lattice > rearrangement? Really hard if not impossible to model... > Don > > Azelio Boriani > > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable level > > for > > that input. I have the 53132A and the 53181A so I can test (with the R&S > > SML01). Is this problem new? Never tested before? > > > > About the OCXO aging: OK, I'm interested in the development of a model > > that > > can be general enough so that it can be "steered", adjusting its > > parameters, during the OCXO life. The data (to adjust the parameters) > > will > > be gathered by the continuous monitoring with the GPS. > > > > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> > > wrote: > > > >> Hi Azelio, > >> > >> Sorry, the counters are 53131A, not 53151A. The A and B inputs are > >> specified > >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, > >> <30 mV > >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. > >> The problem can be at either input alone. The problem is only in the > >> range > >> between 100 MHz and <200 MHz, observed in automatic trigger mode. In > >> manual > >> trigger mode the counter does not trigger at all in that freq range. > >> > >> I cannot tell you an aging model for a particular OCXO, and I doubt if > >> anyone can ;) Aging predictions can only be made on individual OCXO, > >> unit > >> by > >> unit, and the prediction is only valid over a rather limited time. > >> Therefore > >> you must continuously monitor and update your prediction. > >> > >> Best regards > >> > >> Bernd > >> DK1AG > >> > >> -----Ursprüngliche Nachricht----- > >> Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im > >> Auftrag von Azelio Boriani > >> Gesendet: Freitag, 24. August 2012 17:28 > >> An: Discussion of precise time and frequency measurement > >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems > >> > >> From the datasheet: > >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, > >> -20dBm > >> the problem is on A input only? Maybe 150MHz for the A input, being > >> beyond > >> the official specifications, requires more signal on one counter than > >> the > >> other. In the past they all were fine? > >> I have read your paper on the correlation between the real and > >> predicted > >> aging of crystal oscillators. I'm trying to find a suitable model for > >> the > >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman > >> filter. > >> > >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> > >> wrote: > >> > >> > Hi all, > >> > > >> > I have several HP/Agilent frequency counters 53151A (and 53152A) in > >> my > >> lab. > >> > Several of them meanwhile show trigger problems. > >> > > >> > I have tested the A and B channel between 5 MHz and 200 MHZ at three > >> > input levels of -10 dBm, 0 dBm and +10 dBm. > >> > > >> > Those with trigger problems show erroneous numbers at -10 dBm input > >> > level rather selectively, i.e. mostly at 100 MHz, some also at 150 > >> > MHz, while the are o.k. > >> > > >> > Has anyone on this list experienced similar defects and can give a > >> > recommendation how to fix/repair them? > >> > > >> > > >> > > >> > Best regards > >> > > >> > > >> > > >> > Bernd > >> > > >> > DK1AG > >> > > >> > www.axtal.com > >> > > >> > _______________________________________________ > >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> > and follow the instructions there. > >> > > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > > -- > "Neither the voice of authority nor the weight of reason and argument > are as significant as experiment, for thence comes quiet to the mind." > R. Bacon > "If you don't know what it is, don't poke it." > Ghost in the Shell > > > Dr. Don Latham AJ7LL > Six Mile Systems LLP > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > www.lightningforensics.com > www.sixmilesystems.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
DL
Don Latham
Mon, Aug 27, 2012 10:15 PM

There are ways of generating models, such as ARIMA, using past behavior
and the models used to tweak filters and so on. The shifts would have to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for a
given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO taking the
action from a suitable model should help in speeding up the adjustments
and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable

level

for
that input. I have the 53132A and the 53181A so I can test (with the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a

model

that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de
wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I doubt

if

anyone can ;) Aging predictions can only be made on individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz,
-20dBm
the problem is on A input only? Maybe 150MHz for the A input, being
beyond
the official specifications, requires more signal on one counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman
filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de
wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

There are ways of generating models, such as ARIMA, using past behavior and the models used to tweak filters and so on. The shifts would have to be accounted for as they occur, seems to me. Now the interesting possibility is that a shift does not alter the underlying model for a given crystal???? Has to be tested. Don Azelio Boriani > So it can't be done... the possibility to discipline an OCXO taking the > action from a suitable model should help in speeding up the adjustments > and > avoiding the humps as seen in the Allan deviation plot. > > On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: > >> Isn't crystal aging marked by random jumps? perhaps lattice >> rearrangement? Really hard if not impossible to model... >> Don >> >> Azelio Boriani >> > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable >> level >> > for >> > that input. I have the 53132A and the 53181A so I can test (with the >> R&S >> > SML01). Is this problem new? Never tested before? >> > >> > About the OCXO aging: OK, I'm interested in the development of a >> model >> > that >> > can be general enough so that it can be "steered", adjusting its >> > parameters, during the OCXO life. The data (to adjust the >> parameters) >> > will >> > be gathered by the continuous monitoring with the GPS. >> > >> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> >> > wrote: >> > >> >> Hi Azelio, >> >> >> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs are >> >> specified >> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 >> MHz, >> >> <30 mV >> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. >> >> The problem can be at either input alone. The problem is only in >> the >> >> range >> >> between 100 MHz and <200 MHz, observed in automatic trigger mode. >> In >> >> manual >> >> trigger mode the counter does not trigger at all in that freq >> range. >> >> >> >> I cannot tell you an aging model for a particular OCXO, and I doubt >> if >> >> anyone can ;) Aging predictions can only be made on individual >> OCXO, >> >> unit >> >> by >> >> unit, and the prediction is only valid over a rather limited time. >> >> Therefore >> >> you must continuously monitor and update your prediction. >> >> >> >> Best regards >> >> >> >> Bernd >> >> DK1AG >> >> >> >> -----Ursprüngliche Nachricht----- >> >> Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] >> Im >> >> Auftrag von Azelio Boriani >> >> Gesendet: Freitag, 24. August 2012 17:28 >> >> An: Discussion of precise time and frequency measurement >> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems >> >> >> >> From the datasheet: >> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, >> >> -20dBm >> >> the problem is on A input only? Maybe 150MHz for the A input, being >> >> beyond >> >> the official specifications, requires more signal on one counter >> than >> >> the >> >> other. In the past they all were fine? >> >> I have read your paper on the correlation between the real and >> >> predicted >> >> aging of crystal oscillators. I'm trying to find a suitable model >> for >> >> the >> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman >> >> filter. >> >> >> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> >> >> wrote: >> >> >> >> > Hi all, >> >> > >> >> > I have several HP/Agilent frequency counters 53151A (and 53152A) >> in >> >> my >> >> lab. >> >> > Several of them meanwhile show trigger problems. >> >> > >> >> > I have tested the A and B channel between 5 MHz and 200 MHZ at >> three >> >> > input levels of -10 dBm, 0 dBm and +10 dBm. >> >> > >> >> > Those with trigger problems show erroneous numbers at -10 dBm >> input >> >> > level rather selectively, i.e. mostly at 100 MHz, some also at >> 150 >> >> > MHz, while the are o.k. >> >> > >> >> > Has anyone on this list experienced similar defects and can give >> a >> >> > recommendation how to fix/repair them? >> >> > >> >> > >> >> > >> >> > Best regards >> >> > >> >> > >> >> > >> >> > Bernd >> >> > >> >> > DK1AG >> >> > >> >> > www.axtal.com >> >> > >> >> > _______________________________________________ >> >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go >> to >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> > and follow the instructions there. >> >> > >> >> _______________________________________________ >> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> and follow the instructions there. >> >> >> >> >> >> _______________________________________________ >> >> time-nuts mailing list -- time-nuts@febo.com >> >> To unsubscribe, go to >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> and follow the instructions there. >> >> >> > _______________________________________________ >> > time-nuts mailing list -- time-nuts@febo.com >> > To unsubscribe, go to >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> > and follow the instructions there. >> > >> > >> >> >> -- >> "Neither the voice of authority nor the weight of reason and argument >> are as significant as experiment, for thence comes quiet to the mind." >> R. Bacon >> "If you don't know what it is, don't poke it." >> Ghost in the Shell >> >> >> Dr. Don Latham AJ7LL >> Six Mile Systems LLP >> 17850 Six Mile Road >> POB 134 >> Huson, MT, 59846 >> VOX 406-626-4304 >> www.lightningforensics.com >> www.sixmilesystems.com >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." R. Bacon "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
MB
Michael Blazer
Mon, Aug 27, 2012 11:06 PM

Bernd,
Do the 53131A have rear inputs?  At work, we're on contract to
repair 53132As and found the rear input seriously degrades the
performance.  The basic sensitivity (100-200 MHz) is 30 mVrms.  The Rear
input is in parallel with the front panel and the sensitivity is 75
mVrms (full frequency range).
In auto, the counter will attempt to set the trigger level between
the positive and negative peaks.  In manual mode, you need to set the
trigger level and it might be misaligned.

Mike

On 8/27/2012 12:01 PM, Bernd Neubig wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, <30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in the range
between 100 MHz and <200 MHz, observed in automatic trigger mode. In manual
trigger mode the counter does not trigger at all in that freq range.

I cannot tell you an aging model for a particular OCXO, and I doubt if
anyone can ;) Aging predictions can only be made on individual OCXO, unit by
unit, and the prediction is only valid over a rather limited time. Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:

A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm
the problem is on A input only? Maybe 150MHz for the A input, being beyond
the official specifications, requires more signal on one counter than the
other. In the past they all were fine?
I have read your paper on the correlation between the real and predicted
aging of crystal oscillators. I'm trying to find a suitable model for the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A) in my

lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at three
input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm input
level rather selectively, i.e. mostly at 100 MHz, some also at 150
MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give a
recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bernd, Do the 53131A have rear inputs? At work, we're on contract to repair 53132As and found the rear input seriously degrades the performance. The basic sensitivity (100-200 MHz) is 30 mVrms. The Rear input is in parallel with the front panel and the sensitivity is 75 mVrms (full frequency range). In auto, the counter will attempt to set the trigger level between the positive and negative peaks. In manual mode, you need to set the trigger level and it might be misaligned. Mike On 8/27/2012 12:01 PM, Bernd Neubig wrote: > Hi Azelio, > > Sorry, the counters are 53131A, not 53151A. The A and B inputs are specified > from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 MHz, <30 mV > (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. > The problem can be at either input alone. The problem is only in the range > between 100 MHz and <200 MHz, observed in automatic trigger mode. In manual > trigger mode the counter does not trigger at all in that freq range. > > I cannot tell you an aging model for a particular OCXO, and I doubt if > anyone can ;) Aging predictions can only be made on individual OCXO, unit by > unit, and the prediction is only valid over a rather limited time. Therefore > you must continuously monitor and update your prediction. > > Best regards > > Bernd > DK1AG > > -----Ursprüngliche Nachricht----- > Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im > Auftrag von Azelio Boriani > Gesendet: Freitag, 24. August 2012 17:28 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems > > >From the datasheet: > A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, -20dBm > the problem is on A input only? Maybe 150MHz for the A input, being beyond > the official specifications, requires more signal on one counter than the > other. In the past they all were fine? > I have read your paper on the correlation between the real and predicted > aging of crystal oscillators. I'm trying to find a suitable model for the > Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman filter. > > On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> wrote: > >> Hi all, >> >> I have several HP/Agilent frequency counters 53151A (and 53152A) in my > lab. >> Several of them meanwhile show trigger problems. >> >> I have tested the A and B channel between 5 MHz and 200 MHZ at three >> input levels of -10 dBm, 0 dBm and +10 dBm. >> >> Those with trigger problems show erroneous numbers at -10 dBm input >> level rather selectively, i.e. mostly at 100 MHz, some also at 150 >> MHz, while the are o.k. >> >> Has anyone on this list experienced similar defects and can give a >> recommendation how to fix/repair them? >> >> >> >> Best regards >> >> >> >> Bernd >> >> DK1AG >> >> www.axtal.com >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
AB
Azelio Boriani
Mon, Aug 27, 2012 11:09 PM

My starting point is the paper by Neubig on the real vs. predicted aging:
experimenting with the model and try to cut it around the Morion or the
Oscilloquartz OCXOs.

On Tue, Aug 28, 2012 at 12:15 AM, Don Latham djl@montana.com wrote:

There are ways of generating models, such as ARIMA, using past behavior
and the models used to tweak filters and so on. The shifts would have to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for a
given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO taking the
action from a suitable model should help in speeding up the adjustments
and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable

level

for
that input. I have the 53132A and the 53181A so I can test (with the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a

model

that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de
wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I doubt

if

anyone can ;) Aging predictions can only be made on individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz,
-20dBm
the problem is on A input only? Maybe 150MHz for the A input, being
beyond
the official specifications, requires more signal on one counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman
filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de
wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

My starting point is the paper by Neubig on the real vs. predicted aging: experimenting with the model and try to cut it around the Morion or the Oscilloquartz OCXOs. On Tue, Aug 28, 2012 at 12:15 AM, Don Latham <djl@montana.com> wrote: > There are ways of generating models, such as ARIMA, using past behavior > and the models used to tweak filters and so on. The shifts would have to > be accounted for as they occur, seems to me. Now the interesting > possibility is that a shift does not alter the underlying model for a > given crystal???? > Has to be tested. > Don > > Azelio Boriani > > So it can't be done... the possibility to discipline an OCXO taking the > > action from a suitable model should help in speeding up the adjustments > > and > > avoiding the humps as seen in the Allan deviation plot. > > > > On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: > > > >> Isn't crystal aging marked by random jumps? perhaps lattice > >> rearrangement? Really hard if not impossible to model... > >> Don > >> > >> Azelio Boriani > >> > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable > >> level > >> > for > >> > that input. I have the 53132A and the 53181A so I can test (with the > >> R&S > >> > SML01). Is this problem new? Never tested before? > >> > > >> > About the OCXO aging: OK, I'm interested in the development of a > >> model > >> > that > >> > can be general enough so that it can be "steered", adjusting its > >> > parameters, during the OCXO life. The data (to adjust the > >> parameters) > >> > will > >> > be gathered by the continuous monitoring with the GPS. > >> > > >> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> > >> > wrote: > >> > > >> >> Hi Azelio, > >> >> > >> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs are > >> >> specified > >> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 > >> MHz, > >> >> <30 mV > >> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. > >> >> The problem can be at either input alone. The problem is only in > >> the > >> >> range > >> >> between 100 MHz and <200 MHz, observed in automatic trigger mode. > >> In > >> >> manual > >> >> trigger mode the counter does not trigger at all in that freq > >> range. > >> >> > >> >> I cannot tell you an aging model for a particular OCXO, and I doubt > >> if > >> >> anyone can ;) Aging predictions can only be made on individual > >> OCXO, > >> >> unit > >> >> by > >> >> unit, and the prediction is only valid over a rather limited time. > >> >> Therefore > >> >> you must continuously monitor and update your prediction. > >> >> > >> >> Best regards > >> >> > >> >> Bernd > >> >> DK1AG > >> >> > >> >> -----Ursprüngliche Nachricht----- > >> >> Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] > >> Im > >> >> Auftrag von Azelio Boriani > >> >> Gesendet: Freitag, 24. August 2012 17:28 > >> >> An: Discussion of precise time and frequency measurement > >> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems > >> >> > >> >> From the datasheet: > >> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, > >> >> -20dBm > >> >> the problem is on A input only? Maybe 150MHz for the A input, being > >> >> beyond > >> >> the official specifications, requires more signal on one counter > >> than > >> >> the > >> >> other. In the past they all were fine? > >> >> I have read your paper on the correlation between the real and > >> >> predicted > >> >> aging of crystal oscillators. I'm trying to find a suitable model > >> for > >> >> the > >> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman > >> >> filter. > >> >> > >> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> > >> >> wrote: > >> >> > >> >> > Hi all, > >> >> > > >> >> > I have several HP/Agilent frequency counters 53151A (and 53152A) > >> in > >> >> my > >> >> lab. > >> >> > Several of them meanwhile show trigger problems. > >> >> > > >> >> > I have tested the A and B channel between 5 MHz and 200 MHZ at > >> three > >> >> > input levels of -10 dBm, 0 dBm and +10 dBm. > >> >> > > >> >> > Those with trigger problems show erroneous numbers at -10 dBm > >> input > >> >> > level rather selectively, i.e. mostly at 100 MHz, some also at > >> 150 > >> >> > MHz, while the are o.k. > >> >> > > >> >> > Has anyone on this list experienced similar defects and can give > >> a > >> >> > recommendation how to fix/repair them? > >> >> > > >> >> > > >> >> > > >> >> > Best regards > >> >> > > >> >> > > >> >> > > >> >> > Bernd > >> >> > > >> >> > DK1AG > >> >> > > >> >> > www.axtal.com > >> >> > > >> >> > _______________________________________________ > >> >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go > >> to > >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> > and follow the instructions there. > >> >> > > >> >> _______________________________________________ > >> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> and follow the instructions there. > >> >> > >> >> > >> >> _______________________________________________ > >> >> time-nuts mailing list -- time-nuts@febo.com > >> >> To unsubscribe, go to > >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> and follow the instructions there. > >> >> > >> > _______________________________________________ > >> > time-nuts mailing list -- time-nuts@febo.com > >> > To unsubscribe, go to > >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> > and follow the instructions there. > >> > > >> > > >> > >> > >> -- > >> "Neither the voice of authority nor the weight of reason and argument > >> are as significant as experiment, for thence comes quiet to the mind." > >> R. Bacon > >> "If you don't know what it is, don't poke it." > >> Ghost in the Shell > >> > >> > >> Dr. Don Latham AJ7LL > >> Six Mile Systems LLP > >> 17850 Six Mile Road > >> POB 134 > >> Huson, MT, 59846 > >> VOX 406-626-4304 > >> www.lightningforensics.com > >> www.sixmilesystems.com > >> > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > > -- > "Neither the voice of authority nor the weight of reason and argument > are as significant as experiment, for thence comes quiet to the mind." > R. Bacon > "If you don't know what it is, don't poke it." > Ghost in the Shell > > > Dr. Don Latham AJ7LL > Six Mile Systems LLP > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > www.lightningforensics.com > www.sixmilesystems.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
RK
Rick Karlquist
Mon, Aug 27, 2012 11:15 PM

Several decades ago, the concept of the "smart clock" arose
at what was then HP.  The idea was as discussed here to
"characterize" past aging, "predict" future aging, and
then "correct" the aging.  The goal wasn't to turn a quartz
oscillator into an atomic clock replacement, but simply
to get the oscillator through a 1 hour or so "holdover" time
during GPS outages.  It sort of worked for that very limited
purpose, but in general, past performance of HP crystals wasn't
a very good predictor of future results.  Crystals would age
in one direction for a while and possibly slow down as time
when on, but then then might start aging in the other direction.
There were also frequency jumps that were substantial and totally
random.  The reason why the HP crystals were unpredictable was
that all the deterministic processes such as mass preferentially
depositing on the crystal, so as to make the frequency age
lower, had been eliminated by years of manufacturing improvements.
The remaining processes were of the nature of quartz stress
relaxation that were very random.

Rick Karlquist

Don Latham wrote:

There are ways of generating models, such as ARIMA, using past behavior
and the models used to tweak filters and so on. The shifts would have to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for a
given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO taking the
action from a suitable model should help in speeding up the adjustments
and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable

level

for
that input. I have the 53132A and the 53181A so I can test (with the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a

model

that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de
wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I doubt

if

anyone can ;) Aging predictions can only be made on individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz,
-20dBm
the problem is on A input only? Maybe 150MHz for the A input, being
beyond
the official specifications, requires more signal on one counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman
filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de
wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Several decades ago, the concept of the "smart clock" arose at what was then HP. The idea was as discussed here to "characterize" past aging, "predict" future aging, and then "correct" the aging. The goal wasn't to turn a quartz oscillator into an atomic clock replacement, but simply to get the oscillator through a 1 hour or so "holdover" time during GPS outages. It sort of worked for that very limited purpose, but in general, past performance of HP crystals wasn't a very good predictor of future results. Crystals would age in one direction for a while and possibly slow down as time when on, but then then might start aging in the other direction. There were also frequency jumps that were substantial and totally random. The reason why the HP crystals were unpredictable was that all the deterministic processes such as mass preferentially depositing on the crystal, so as to make the frequency age lower, had been eliminated by years of manufacturing improvements. The remaining processes were of the nature of quartz stress relaxation that were very random. Rick Karlquist Don Latham wrote: > There are ways of generating models, such as ARIMA, using past behavior > and the models used to tweak filters and so on. The shifts would have to > be accounted for as they occur, seems to me. Now the interesting > possibility is that a shift does not alter the underlying model for a > given crystal???? > Has to be tested. > Don > > Azelio Boriani >> So it can't be done... the possibility to discipline an OCXO taking the >> action from a suitable model should help in speeding up the adjustments >> and >> avoiding the humps as seen in the Allan deviation plot. >> >> On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: >> >>> Isn't crystal aging marked by random jumps? perhaps lattice >>> rearrangement? Really hard if not impossible to model... >>> Don >>> >>> Azelio Boriani >>> > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable >>> level >>> > for >>> > that input. I have the 53132A and the 53181A so I can test (with the >>> R&S >>> > SML01). Is this problem new? Never tested before? >>> > >>> > About the OCXO aging: OK, I'm interested in the development of a >>> model >>> > that >>> > can be general enough so that it can be "steered", adjusting its >>> > parameters, during the OCXO life. The data (to adjust the >>> parameters) >>> > will >>> > be gathered by the continuous monitoring with the GPS. >>> > >>> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> >>> > wrote: >>> > >>> >> Hi Azelio, >>> >> >>> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs are >>> >> specified >>> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 >>> MHz, >>> >> <30 mV >>> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. >>> >> The problem can be at either input alone. The problem is only in >>> the >>> >> range >>> >> between 100 MHz and <200 MHz, observed in automatic trigger mode. >>> In >>> >> manual >>> >> trigger mode the counter does not trigger at all in that freq >>> range. >>> >> >>> >> I cannot tell you an aging model for a particular OCXO, and I doubt >>> if >>> >> anyone can ;) Aging predictions can only be made on individual >>> OCXO, >>> >> unit >>> >> by >>> >> unit, and the prediction is only valid over a rather limited time. >>> >> Therefore >>> >> you must continuously monitor and update your prediction. >>> >> >>> >> Best regards >>> >> >>> >> Bernd >>> >> DK1AG >>> >> >>> >> -----Ursprüngliche Nachricht----- >>> >> Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] >>> Im >>> >> Auftrag von Azelio Boriani >>> >> Gesendet: Freitag, 24. August 2012 17:28 >>> >> An: Discussion of precise time and frequency measurement >>> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems >>> >> >>> >> From the datasheet: >>> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, >>> >> -20dBm >>> >> the problem is on A input only? Maybe 150MHz for the A input, being >>> >> beyond >>> >> the official specifications, requires more signal on one counter >>> than >>> >> the >>> >> other. In the past they all were fine? >>> >> I have read your paper on the correlation between the real and >>> >> predicted >>> >> aging of crystal oscillators. I'm trying to find a suitable model >>> for >>> >> the >>> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman >>> >> filter. >>> >> >>> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> >>> >> wrote: >>> >> >>> >> > Hi all, >>> >> > >>> >> > I have several HP/Agilent frequency counters 53151A (and 53152A) >>> in >>> >> my >>> >> lab. >>> >> > Several of them meanwhile show trigger problems. >>> >> > >>> >> > I have tested the A and B channel between 5 MHz and 200 MHZ at >>> three >>> >> > input levels of -10 dBm, 0 dBm and +10 dBm. >>> >> > >>> >> > Those with trigger problems show erroneous numbers at -10 dBm >>> input >>> >> > level rather selectively, i.e. mostly at 100 MHz, some also at >>> 150 >>> >> > MHz, while the are o.k. >>> >> > >>> >> > Has anyone on this list experienced similar defects and can give >>> a >>> >> > recommendation how to fix/repair them? >>> >> > >>> >> > >>> >> > >>> >> > Best regards >>> >> > >>> >> > >>> >> > >>> >> > Bernd >>> >> > >>> >> > DK1AG >>> >> > >>> >> > www.axtal.com >>> >> > >>> >> > _______________________________________________ >>> >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go >>> to >>> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >> > and follow the instructions there. >>> >> > >>> >> _______________________________________________ >>> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >> and follow the instructions there. >>> >> >>> >> >>> >> _______________________________________________ >>> >> time-nuts mailing list -- time-nuts@febo.com >>> >> To unsubscribe, go to >>> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >> and follow the instructions there. >>> >> >>> > _______________________________________________ >>> > time-nuts mailing list -- time-nuts@febo.com >>> > To unsubscribe, go to >>> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> > and follow the instructions there. >>> > >>> > >>> >>> >>> -- >>> "Neither the voice of authority nor the weight of reason and argument >>> are as significant as experiment, for thence comes quiet to the mind." >>> R. Bacon >>> "If you don't know what it is, don't poke it." >>> Ghost in the Shell >>> >>> >>> Dr. Don Latham AJ7LL >>> Six Mile Systems LLP >>> 17850 Six Mile Road >>> POB 134 >>> Huson, MT, 59846 >>> VOX 406-626-4304 >>> www.lightningforensics.com >>> www.sixmilesystems.com >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> > > > -- > "Neither the voice of authority nor the weight of reason and argument > are as significant as experiment, for thence comes quiet to the mind." > R. Bacon > "If you don't know what it is, don't poke it." > Ghost in the Shell > > > Dr. Don Latham AJ7LL > Six Mile Systems LLP > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > www.lightningforensics.com > www.sixmilesystems.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
DL
Don Latham
Tue, Aug 28, 2012 12:32 AM

Could you refresh me with the reference or email the paper? I'd like to
give it a read.
Don

Azelio Boriani

My starting point is the paper by Neubig on the real vs. predicted
aging:
experimenting with the model and try to cut it around the Morion or the
Oscilloquartz OCXOs.

On Tue, Aug 28, 2012 at 12:15 AM, Don Latham djl@montana.com wrote:

There are ways of generating models, such as ARIMA, using past
behavior
and the models used to tweak filters and so on. The shifts would have
to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for a
given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO taking

the

action from a suitable model should help in speeding up the

adjustments

and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable

level

for
that input. I have the 53132A and the 53181A so I can test (with

the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a

model

that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig

wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs

are

specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger

mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I

doubt

if

anyone can ;) Aging predictions can only be made on individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited

time.

Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger

problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min

50MHz,

-20dBm
the problem is on A input only? Maybe 150MHz for the A input,

being

beyond
the official specifications, requires more signal on one counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable

model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a

Kalman

filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig

wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and

53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can

give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe,

go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go

to

https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and

argument

are as significant as experiment, for thence comes quiet to the

mind."

R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

Could you refresh me with the reference or email the paper? I'd like to give it a read. Don Azelio Boriani > My starting point is the paper by Neubig on the real vs. predicted > aging: > experimenting with the model and try to cut it around the Morion or the > Oscilloquartz OCXOs. > > On Tue, Aug 28, 2012 at 12:15 AM, Don Latham <djl@montana.com> wrote: > >> There are ways of generating models, such as ARIMA, using past >> behavior >> and the models used to tweak filters and so on. The shifts would have >> to >> be accounted for as they occur, seems to me. Now the interesting >> possibility is that a shift does not alter the underlying model for a >> given crystal???? >> Has to be tested. >> Don >> >> Azelio Boriani >> > So it can't be done... the possibility to discipline an OCXO taking >> the >> > action from a suitable model should help in speeding up the >> adjustments >> > and >> > avoiding the humps as seen in the Allan deviation plot. >> > >> > On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: >> > >> >> Isn't crystal aging marked by random jumps? perhaps lattice >> >> rearrangement? Really hard if not impossible to model... >> >> Don >> >> >> >> Azelio Boriani >> >> > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable >> >> level >> >> > for >> >> > that input. I have the 53132A and the 53181A so I can test (with >> the >> >> R&S >> >> > SML01). Is this problem new? Never tested before? >> >> > >> >> > About the OCXO aging: OK, I'm interested in the development of a >> >> model >> >> > that >> >> > can be general enough so that it can be "steered", adjusting its >> >> > parameters, during the OCXO life. The data (to adjust the >> >> parameters) >> >> > will >> >> > be gathered by the continuous monitoring with the GPS. >> >> > >> >> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig >> <bneubig@t-online.de> >> >> > wrote: >> >> > >> >> >> Hi Azelio, >> >> >> >> >> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs >> are >> >> >> specified >> >> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 >> >> MHz, >> >> >> <30 mV >> >> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. >> >> >> The problem can be at either input alone. The problem is only in >> >> the >> >> >> range >> >> >> between 100 MHz and <200 MHz, observed in automatic trigger >> mode. >> >> In >> >> >> manual >> >> >> trigger mode the counter does not trigger at all in that freq >> >> range. >> >> >> >> >> >> I cannot tell you an aging model for a particular OCXO, and I >> doubt >> >> if >> >> >> anyone can ;) Aging predictions can only be made on individual >> >> OCXO, >> >> >> unit >> >> >> by >> >> >> unit, and the prediction is only valid over a rather limited >> time. >> >> >> Therefore >> >> >> you must continuously monitor and update your prediction. >> >> >> >> >> >> Best regards >> >> >> >> >> >> Bernd >> >> >> DK1AG >> >> >> >> >> >> -----Ursprüngliche Nachricht----- >> >> >> Von: time-nuts-bounces@febo.com >> [mailto:time-nuts-bounces@febo.com] >> >> Im >> >> >> Auftrag von Azelio Boriani >> >> >> Gesendet: Freitag, 24. August 2012 17:28 >> >> >> An: Discussion of precise time and frequency measurement >> >> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger >> problems >> >> >> >> >> >> From the datasheet: >> >> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min >> 50MHz, >> >> >> -20dBm >> >> >> the problem is on A input only? Maybe 150MHz for the A input, >> being >> >> >> beyond >> >> >> the official specifications, requires more signal on one counter >> >> than >> >> >> the >> >> >> other. In the past they all were fine? >> >> >> I have read your paper on the correlation between the real and >> >> >> predicted >> >> >> aging of crystal oscillators. I'm trying to find a suitable >> model >> >> for >> >> >> the >> >> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a >> Kalman >> >> >> filter. >> >> >> >> >> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig >> <BNeubig@t-online.de> >> >> >> wrote: >> >> >> >> >> >> > Hi all, >> >> >> > >> >> >> > I have several HP/Agilent frequency counters 53151A (and >> 53152A) >> >> in >> >> >> my >> >> >> lab. >> >> >> > Several of them meanwhile show trigger problems. >> >> >> > >> >> >> > I have tested the A and B channel between 5 MHz and 200 MHZ at >> >> three >> >> >> > input levels of -10 dBm, 0 dBm and +10 dBm. >> >> >> > >> >> >> > Those with trigger problems show erroneous numbers at -10 dBm >> >> input >> >> >> > level rather selectively, i.e. mostly at 100 MHz, some also at >> >> 150 >> >> >> > MHz, while the are o.k. >> >> >> > >> >> >> > Has anyone on this list experienced similar defects and can >> give >> >> a >> >> >> > recommendation how to fix/repair them? >> >> >> > >> >> >> > >> >> >> > >> >> >> > Best regards >> >> >> > >> >> >> > >> >> >> > >> >> >> > Bernd >> >> >> > >> >> >> > DK1AG >> >> >> > >> >> >> > www.axtal.com >> >> >> > >> >> >> > _______________________________________________ >> >> >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, >> go >> >> to >> >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> > and follow the instructions there. >> >> >> > >> >> >> _______________________________________________ >> >> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go >> to >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> and follow the instructions there. >> >> >> >> >> >> >> >> >> _______________________________________________ >> >> >> time-nuts mailing list -- time-nuts@febo.com >> >> >> To unsubscribe, go to >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> and follow the instructions there. >> >> >> >> >> > _______________________________________________ >> >> > time-nuts mailing list -- time-nuts@febo.com >> >> > To unsubscribe, go to >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> > and follow the instructions there. >> >> > >> >> > >> >> >> >> >> >> -- >> >> "Neither the voice of authority nor the weight of reason and >> argument >> >> are as significant as experiment, for thence comes quiet to the >> mind." >> >> R. Bacon >> >> "If you don't know what it is, don't poke it." >> >> Ghost in the Shell >> >> >> >> >> >> Dr. Don Latham AJ7LL >> >> Six Mile Systems LLP >> >> 17850 Six Mile Road >> >> POB 134 >> >> Huson, MT, 59846 >> >> VOX 406-626-4304 >> >> www.lightningforensics.com >> >> www.sixmilesystems.com >> >> >> >> >> >> >> >> _______________________________________________ >> >> time-nuts mailing list -- time-nuts@febo.com >> >> To unsubscribe, go to >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> and follow the instructions there. >> >> >> > _______________________________________________ >> > time-nuts mailing list -- time-nuts@febo.com >> > To unsubscribe, go to >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> > and follow the instructions there. >> > >> > >> >> >> -- >> "Neither the voice of authority nor the weight of reason and argument >> are as significant as experiment, for thence comes quiet to the mind." >> R. Bacon >> "If you don't know what it is, don't poke it." >> Ghost in the Shell >> >> >> Dr. Don Latham AJ7LL >> Six Mile Systems LLP >> 17850 Six Mile Road >> POB 134 >> Huson, MT, 59846 >> VOX 406-626-4304 >> www.lightningforensics.com >> www.sixmilesystems.com >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." R. Bacon "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
JL
Jim Lux
Tue, Aug 28, 2012 2:33 AM

On 8/27/12 4:15 PM, Rick Karlquist wrote:

Several decades ago, the concept of the "smart clock" arose
at what was then HP.  The idea was as discussed here to
"characterize" past aging, "predict" future aging, and
then "correct" the aging.  The goal wasn't to turn a quartz
oscillator into an atomic clock replacement, but simply
to get the oscillator through a 1 hour or so "holdover" time
during GPS outages.  It sort of worked for that very limited
purpose, but in general, past performance of HP crystals wasn't
a very good predictor of future results.  Crystals would age
in one direction for a while and possibly slow down as time
when on, but then then might start aging in the other direction.
There were also frequency jumps that were substantial and totally
random.  The reason why the HP crystals were unpredictable was
that all the deterministic processes such as mass preferentially
depositing on the crystal, so as to make the frequency age
lower, had been eliminated by years of manufacturing improvements.
The remaining processes were of the nature of quartz stress
relaxation that were very random.

Rick Karlquist

We see similar things in USOs  (and other components as well) for
spaceflight..

Things that people worried about 60 or more years ago just don't occur
anymore.  People used to obsessively try to match diodes in HV strings,
for instance, because the process variability was high enough to make a
difference.  These days, you get a reel of diodes and they're all pretty
much the same, and even reel to reel from month to month they don't
change much.    That's what all that 6-sigma stuff is all about, after all.

All the low hanging, and even middle hanging, fruit has been picked..

(one big exception.. ICs which are not designed for radiation tolerance
seem to have large variability in radiation tolerance..it's just not
something that's controlled for in the process)

On 8/27/12 4:15 PM, Rick Karlquist wrote: > Several decades ago, the concept of the "smart clock" arose > at what was then HP. The idea was as discussed here to > "characterize" past aging, "predict" future aging, and > then "correct" the aging. The goal wasn't to turn a quartz > oscillator into an atomic clock replacement, but simply > to get the oscillator through a 1 hour or so "holdover" time > during GPS outages. It sort of worked for that very limited > purpose, but in general, past performance of HP crystals wasn't > a very good predictor of future results. Crystals would age > in one direction for a while and possibly slow down as time > when on, but then then might start aging in the other direction. > There were also frequency jumps that were substantial and totally > random. The reason why the HP crystals were unpredictable was > that all the deterministic processes such as mass preferentially > depositing on the crystal, so as to make the frequency age > lower, had been eliminated by years of manufacturing improvements. > The remaining processes were of the nature of quartz stress > relaxation that were very random. > > Rick Karlquist We see similar things in USOs (and other components as well) for spaceflight.. Things that people worried about 60 or more years ago just don't occur anymore. People used to obsessively try to match diodes in HV strings, for instance, because the process variability was high enough to make a difference. These days, you get a reel of diodes and they're all pretty much the same, and even reel to reel from month to month they don't change much. That's what all that 6-sigma stuff is all about, after all. All the low hanging, and even middle hanging, fruit has been picked.. (one big exception.. ICs which are not designed for radiation tolerance seem to have large variability in radiation tolerance..it's just not something that's controlled for in the process)
BN
Bernd Neubig
Tue, Aug 28, 2012 5:28 AM

Don,

There are two papers authored by me on the subject:

  1. Comparison of Passive and Active Aging Of SC-Cut and AT-Cut Crystals;
    Proceedings IEEE International Frequency Control Symposium
    Hawaii May 1996, pp.316-322
  2. Correlation of predicted and real aging behaviour; Proc. 11th European
    Frequency & Time Forum
    Neuchatel, March 1997, pp.268-272
    Both can be downloaded on the AXTAL website www.axtal.com -> Technical Notes
    -> Technical Articles and Publications -> Quartz crystal units

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Don Latham
Gesendet: Dienstag, 28. August 2012 02:32
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53131A (not 53151A) trigger
problems

Could you refresh me with the reference or email the paper? I'd like to give
it a read.
Don

Don, There are two papers authored by me on the subject: 1. Comparison of Passive and Active Aging Of SC-Cut and AT-Cut Crystals; Proceedings IEEE International Frequency Control Symposium Hawaii May 1996, pp.316-322 2. Correlation of predicted and real aging behaviour; Proc. 11th European Frequency & Time Forum Neuchatel, March 1997, pp.268-272 Both can be downloaded on the AXTAL website www.axtal.com -> Technical Notes -> Technical Articles and Publications -> Quartz crystal units Bernd DK1AG -----Ursprüngliche Nachricht----- Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im Auftrag von Don Latham Gesendet: Dienstag, 28. August 2012 02:32 An: Discussion of precise time and frequency measurement Betreff: Re: [time-nuts] HP/Agilet counter 53131A (not 53151A) trigger problems Could you refresh me with the reference or email the paper? I'd like to give it a read. Don
CA
Chris Albertson
Tue, Aug 28, 2012 5:45 AM

On Mon, Aug 27, 2012 at 7:33 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/27/12 4:15 PM, Rick Karlquist wrote:

Several decades ago, the concept of the "smart clock" arose
at what was then HP.  The idea was as discussed here to
"characterize" past aging, "predict" future aging, and
then "correct" the aging.

We know what a OCXO is and a TCXO is.  I was at a presentation at work a
whike back and they called what you describe a "MPCXO"  or MicroProcessor
Compensated XO.    They said the characteristics were between the OCXO and
TCXO

Doesn't the thunderbolt do this.  I think it watches the aging rate of the
OCXO and adjusts during hold over.

--

Chris Albertson
Redondo Beach, California

On Mon, Aug 27, 2012 at 7:33 PM, Jim Lux <jimlux@earthlink.net> wrote: > On 8/27/12 4:15 PM, Rick Karlquist wrote: > >> Several decades ago, the concept of the "smart clock" arose >> at what was then HP. The idea was as discussed here to >> "characterize" past aging, "predict" future aging, and >> then "correct" the aging. >> > We know what a OCXO is and a TCXO is. I was at a presentation at work a whike back and they called what you describe a "MPCXO" or MicroProcessor Compensated XO. They said the characteristics were between the OCXO and TCXO Doesn't the thunderbolt do this. I think it watches the aging rate of the OCXO and adjusts during hold over. -- Chris Albertson Redondo Beach, California
DL
Don Latham
Tue, Aug 28, 2012 6:06 AM

Thank you very much, Bernd. The papers are on the stack. Now if I can
just get to them, they look very interesting.
73, Don

Bernd Neubig

Don,

There are two papers authored by me on the subject:

  1. Comparison of Passive and Active Aging Of SC-Cut and AT-Cut Crystals;
    Proceedings IEEE International Frequency Control Symposium
    Hawaii May 1996, pp.316-322
  2. Correlation of predicted and real aging behaviour; Proc. 11th
    European
    Frequency & Time Forum
    Neuchatel, March 1997, pp.268-272
    Both can be downloaded on the AXTAL website www.axtal.com -> Technical
    Notes
    -> Technical Articles and Publications -> Quartz crystal units

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im
Auftrag von Don Latham
Gesendet: Dienstag, 28. August 2012 02:32
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53131A (not 53151A) trigger
problems

Could you refresh me with the reference or email the paper? I'd like to
give
it a read.
Don


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
De Erroribus Medicorum, R. Bacon, 13th century.
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

Thank you very much, Bernd. The papers are on the stack. Now if I can just get to them, they look very interesting. 73, Don Bernd Neubig > Don, > > There are two papers authored by me on the subject: > 1. Comparison of Passive and Active Aging Of SC-Cut and AT-Cut Crystals; > Proceedings IEEE International Frequency Control Symposium > Hawaii May 1996, pp.316-322 > 2. Correlation of predicted and real aging behaviour; Proc. 11th > European > Frequency & Time Forum > Neuchatel, March 1997, pp.268-272 > Both can be downloaded on the AXTAL website www.axtal.com -> Technical > Notes > -> Technical Articles and Publications -> Quartz crystal units > > Bernd > DK1AG > > > -----Ursprüngliche Nachricht----- > Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] Im > Auftrag von Don Latham > Gesendet: Dienstag, 28. August 2012 02:32 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] HP/Agilet counter 53131A (not 53151A) trigger > problems > > Could you refresh me with the reference or email the paper? I'd like to > give > it a read. > Don > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." De Erroribus Medicorum, R. Bacon, 13th century. "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
W
WB6BNQ
Tue, Aug 28, 2012 6:45 AM

Chris,

That is a big no !  What the Thunderbolt is doing is adjusting the OCXO to keep
it aligned with a known reference from the GPS system.  The fact that a
microprocessor is involved is from a totally different perspective.

A microprocessor controlled XO is a non oven crystal oscillator system that has
additional computational control providing a bit more than just mere passive
temperature compensation.  The additional computational capability deals with
having coefficients of that particular oscillator's behavior pre coded to
compensate for the nonlinear behavior over a given temperature range.

A microprocessor controlled XO system allows for using cheap crystals with
minimum processing time and costs.  Because of limited storage space there is no
way for the system to have enough data to even try to compete with the quality of
a decent OCXO.  Beyond its initial calibration setup, it has no way of keeping it
tied to a known reference, like the Thunderbolt is doing.

Bill....WB6BNQ

Chris Albertson wrote:

On Mon, Aug 27, 2012 at 7:33 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/27/12 4:15 PM, Rick Karlquist wrote:

Several decades ago, the concept of the "smart clock" arose
at what was then HP.  The idea was as discussed here to
"characterize" past aging, "predict" future aging, and
then "correct" the aging.

We know what a OCXO is and a TCXO is.  I was at a presentation at work a
whike back and they called what you describe a "MPCXO"  or MicroProcessor
Compensated XO.    They said the characteristics were between the OCXO and
TCXO

Doesn't the thunderbolt do this.  I think it watches the aging rate of the
OCXO and adjusts during hold over.

--

Chris Albertson
Redondo Beach, California


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Chris, That is a big no ! What the Thunderbolt is doing is adjusting the OCXO to keep it aligned with a known reference from the GPS system. The fact that a microprocessor is involved is from a totally different perspective. A microprocessor controlled XO is a non oven crystal oscillator system that has additional computational control providing a bit more than just mere passive temperature compensation. The additional computational capability deals with having coefficients of that particular oscillator's behavior pre coded to compensate for the nonlinear behavior over a given temperature range. A microprocessor controlled XO system allows for using cheap crystals with minimum processing time and costs. Because of limited storage space there is no way for the system to have enough data to even try to compete with the quality of a decent OCXO. Beyond its initial calibration setup, it has no way of keeping it tied to a known reference, like the Thunderbolt is doing. Bill....WB6BNQ Chris Albertson wrote: > On Mon, Aug 27, 2012 at 7:33 PM, Jim Lux <jimlux@earthlink.net> wrote: > > > On 8/27/12 4:15 PM, Rick Karlquist wrote: > > > >> Several decades ago, the concept of the "smart clock" arose > >> at what was then HP. The idea was as discussed here to > >> "characterize" past aging, "predict" future aging, and > >> then "correct" the aging. > >> > > > We know what a OCXO is and a TCXO is. I was at a presentation at work a > whike back and they called what you describe a "MPCXO" or MicroProcessor > Compensated XO. They said the characteristics were between the OCXO and > TCXO > > Doesn't the thunderbolt do this. I think it watches the aging rate of the > OCXO and adjusts during hold over. > > -- > > Chris Albertson > Redondo Beach, California > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AB
Azelio Boriani
Tue, Aug 28, 2012 8:12 AM

www.axtal.com/data/publ/aging_e.pdf

On Tue, Aug 28, 2012 at 2:32 AM, Don Latham djl@montana.com wrote:

Could you refresh me with the reference or email the paper? I'd like to
give it a read.
Don

Azelio Boriani

My starting point is the paper by Neubig on the real vs. predicted
aging:
experimenting with the model and try to cut it around the Morion or the
Oscilloquartz OCXOs.

On Tue, Aug 28, 2012 at 12:15 AM, Don Latham djl@montana.com wrote:

There are ways of generating models, such as ARIMA, using past
behavior
and the models used to tweak filters and so on. The shifts would have
to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for a
given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO taking

the

action from a suitable model should help in speeding up the

adjustments

and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable

level

for
that input. I have the 53132A and the 53181A so I can test (with

the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a

model

that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig

wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs

are

specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger

mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I

doubt

if

anyone can ;) Aging predictions can only be made on individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited

time.

Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger

problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min

50MHz,

-20dBm
the problem is on A input only? Maybe 150MHz for the A input,

being

beyond
the official specifications, requires more signal on one counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable

model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a

Kalman

filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig

wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and

53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can

give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe,

go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go

to

https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and

argument

are as significant as experiment, for thence comes quiet to the

mind."

R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

www.*axtal*.com/data/publ/*aging*_e.pdf On Tue, Aug 28, 2012 at 2:32 AM, Don Latham <djl@montana.com> wrote: > Could you refresh me with the reference or email the paper? I'd like to > give it a read. > Don > > Azelio Boriani > > My starting point is the paper by Neubig on the real vs. predicted > > aging: > > experimenting with the model and try to cut it around the Morion or the > > Oscilloquartz OCXOs. > > > > On Tue, Aug 28, 2012 at 12:15 AM, Don Latham <djl@montana.com> wrote: > > > >> There are ways of generating models, such as ARIMA, using past > >> behavior > >> and the models used to tweak filters and so on. The shifts would have > >> to > >> be accounted for as they occur, seems to me. Now the interesting > >> possibility is that a shift does not alter the underlying model for a > >> given crystal???? > >> Has to be tested. > >> Don > >> > >> Azelio Boriani > >> > So it can't be done... the possibility to discipline an OCXO taking > >> the > >> > action from a suitable model should help in speeding up the > >> adjustments > >> > and > >> > avoiding the humps as seen in the Allan deviation plot. > >> > > >> > On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: > >> > > >> >> Isn't crystal aging marked by random jumps? perhaps lattice > >> >> rearrangement? Really hard if not impossible to model... > >> >> Don > >> >> > >> >> Azelio Boriani > >> >> > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable > >> >> level > >> >> > for > >> >> > that input. I have the 53132A and the 53181A so I can test (with > >> the > >> >> R&S > >> >> > SML01). Is this problem new? Never tested before? > >> >> > > >> >> > About the OCXO aging: OK, I'm interested in the development of a > >> >> model > >> >> > that > >> >> > can be general enough so that it can be "steered", adjusting its > >> >> > parameters, during the OCXO life. The data (to adjust the > >> >> parameters) > >> >> > will > >> >> > be gathered by the continuous monitoring with the GPS. > >> >> > > >> >> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig > >> <bneubig@t-online.de> > >> >> > wrote: > >> >> > > >> >> >> Hi Azelio, > >> >> >> > >> >> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs > >> are > >> >> >> specified > >> >> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 > >> >> MHz, > >> >> >> <30 mV > >> >> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. > >> >> >> The problem can be at either input alone. The problem is only in > >> >> the > >> >> >> range > >> >> >> between 100 MHz and <200 MHz, observed in automatic trigger > >> mode. > >> >> In > >> >> >> manual > >> >> >> trigger mode the counter does not trigger at all in that freq > >> >> range. > >> >> >> > >> >> >> I cannot tell you an aging model for a particular OCXO, and I > >> doubt > >> >> if > >> >> >> anyone can ;) Aging predictions can only be made on individual > >> >> OCXO, > >> >> >> unit > >> >> >> by > >> >> >> unit, and the prediction is only valid over a rather limited > >> time. > >> >> >> Therefore > >> >> >> you must continuously monitor and update your prediction. > >> >> >> > >> >> >> Best regards > >> >> >> > >> >> >> Bernd > >> >> >> DK1AG > >> >> >> > >> >> >> -----Ursprüngliche Nachricht----- > >> >> >> Von: time-nuts-bounces@febo.com > >> [mailto:time-nuts-bounces@febo.com] > >> >> Im > >> >> >> Auftrag von Azelio Boriani > >> >> >> Gesendet: Freitag, 24. August 2012 17:28 > >> >> >> An: Discussion of precise time and frequency measurement > >> >> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger > >> problems > >> >> >> > >> >> >> From the datasheet: > >> >> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min > >> 50MHz, > >> >> >> -20dBm > >> >> >> the problem is on A input only? Maybe 150MHz for the A input, > >> being > >> >> >> beyond > >> >> >> the official specifications, requires more signal on one counter > >> >> than > >> >> >> the > >> >> >> other. In the past they all were fine? > >> >> >> I have read your paper on the correlation between the real and > >> >> >> predicted > >> >> >> aging of crystal oscillators. I'm trying to find a suitable > >> model > >> >> for > >> >> >> the > >> >> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a > >> Kalman > >> >> >> filter. > >> >> >> > >> >> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig > >> <BNeubig@t-online.de> > >> >> >> wrote: > >> >> >> > >> >> >> > Hi all, > >> >> >> > > >> >> >> > I have several HP/Agilent frequency counters 53151A (and > >> 53152A) > >> >> in > >> >> >> my > >> >> >> lab. > >> >> >> > Several of them meanwhile show trigger problems. > >> >> >> > > >> >> >> > I have tested the A and B channel between 5 MHz and 200 MHZ at > >> >> three > >> >> >> > input levels of -10 dBm, 0 dBm and +10 dBm. > >> >> >> > > >> >> >> > Those with trigger problems show erroneous numbers at -10 dBm > >> >> input > >> >> >> > level rather selectively, i.e. mostly at 100 MHz, some also at > >> >> 150 > >> >> >> > MHz, while the are o.k. > >> >> >> > > >> >> >> > Has anyone on this list experienced similar defects and can > >> give > >> >> a > >> >> >> > recommendation how to fix/repair them? > >> >> >> > > >> >> >> > > >> >> >> > > >> >> >> > Best regards > >> >> >> > > >> >> >> > > >> >> >> > > >> >> >> > Bernd > >> >> >> > > >> >> >> > DK1AG > >> >> >> > > >> >> >> > www.axtal.com > >> >> >> > > >> >> >> > _______________________________________________ > >> >> >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, > >> go > >> >> to > >> >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> >> > and follow the instructions there. > >> >> >> > > >> >> >> _______________________________________________ > >> >> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go > >> to > >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> >> and follow the instructions there. > >> >> >> > >> >> >> > >> >> >> _______________________________________________ > >> >> >> time-nuts mailing list -- time-nuts@febo.com > >> >> >> To unsubscribe, go to > >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> >> and follow the instructions there. > >> >> >> > >> >> > _______________________________________________ > >> >> > time-nuts mailing list -- time-nuts@febo.com > >> >> > To unsubscribe, go to > >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> > and follow the instructions there. > >> >> > > >> >> > > >> >> > >> >> > >> >> -- > >> >> "Neither the voice of authority nor the weight of reason and > >> argument > >> >> are as significant as experiment, for thence comes quiet to the > >> mind." > >> >> R. Bacon > >> >> "If you don't know what it is, don't poke it." > >> >> Ghost in the Shell > >> >> > >> >> > >> >> Dr. Don Latham AJ7LL > >> >> Six Mile Systems LLP > >> >> 17850 Six Mile Road > >> >> POB 134 > >> >> Huson, MT, 59846 > >> >> VOX 406-626-4304 > >> >> www.lightningforensics.com > >> >> www.sixmilesystems.com > >> >> > >> >> > >> >> > >> >> _______________________________________________ > >> >> time-nuts mailing list -- time-nuts@febo.com > >> >> To unsubscribe, go to > >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> >> and follow the instructions there. > >> >> > >> > _______________________________________________ > >> > time-nuts mailing list -- time-nuts@febo.com > >> > To unsubscribe, go to > >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> > and follow the instructions there. > >> > > >> > > >> > >> > >> -- > >> "Neither the voice of authority nor the weight of reason and argument > >> are as significant as experiment, for thence comes quiet to the mind." > >> R. Bacon > >> "If you don't know what it is, don't poke it." > >> Ghost in the Shell > >> > >> > >> Dr. Don Latham AJ7LL > >> Six Mile Systems LLP > >> 17850 Six Mile Road > >> POB 134 > >> Huson, MT, 59846 > >> VOX 406-626-4304 > >> www.lightningforensics.com > >> www.sixmilesystems.com > >> > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > > -- > "Neither the voice of authority nor the weight of reason and argument > are as significant as experiment, for thence comes quiet to the mind." > R. Bacon > "If you don't know what it is, don't poke it." > Ghost in the Shell > > > Dr. Don Latham AJ7LL > Six Mile Systems LLP > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > www.lightningforensics.com > www.sixmilesystems.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
JL
Jim Lux
Tue, Aug 28, 2012 1:17 PM

On 8/27/12 10:45 PM, Chris Albertson wrote:

On Mon, Aug 27, 2012 at 7:33 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/27/12 4:15 PM, Rick Karlquist wrote:

Several decades ago, the concept of the "smart clock" arose
at what was then HP.  The idea was as discussed here to
"characterize" past aging, "predict" future aging, and
then "correct" the aging.

We know what a OCXO is and a TCXO is.  I was at a presentation at work a
whike back and they called what you describe a "MPCXO"  or MicroProcessor
Compensated XO.    They said the characteristics were between the OCXO and
TCXO

Doesn't the thunderbolt do this.  I think it watches the aging rate of the
OCXO and adjusts during hold over.

Not really.  The MCXO has a one time calibration for frequency vs
temperature that's programmed into it (and some use very clever ways
to measure the temperature).  The disciplining algorithms in a GPSDO are
a bit smarter; some explicitly develop a model for the f vs T and apply
it, in others it's essentially embedded in a higher order filter which
takes the measured T, along with other parameters, into the filter.

I don't know if the GPSDOs try to do a time series fit/model (at least
for low order terms) to deal with things like diurnal variation.  They
could.

I should note that the MCXO approach, popularized as a "better TCXO" in
a small low power package, is also used in some software defined radios,
except that there's no separate microcontroller.  The frequency vs
temperature characteristic is just embedded in the other algorithms in
the radio's host processor.

On 8/27/12 10:45 PM, Chris Albertson wrote: > On Mon, Aug 27, 2012 at 7:33 PM, Jim Lux <jimlux@earthlink.net> wrote: > >> On 8/27/12 4:15 PM, Rick Karlquist wrote: >> >>> Several decades ago, the concept of the "smart clock" arose >>> at what was then HP. The idea was as discussed here to >>> "characterize" past aging, "predict" future aging, and >>> then "correct" the aging. >>> >> > We know what a OCXO is and a TCXO is. I was at a presentation at work a > whike back and they called what you describe a "MPCXO" or MicroProcessor > Compensated XO. They said the characteristics were between the OCXO and > TCXO > > Doesn't the thunderbolt do this. I think it watches the aging rate of the > OCXO and adjusts during hold over. > Not really. The MCXO has a one time calibration for frequency vs temperature that's programmed into it (and some use *very* clever ways to measure the temperature). The disciplining algorithms in a GPSDO are a bit smarter; some explicitly develop a model for the f vs T and apply it, in others it's essentially embedded in a higher order filter which takes the measured T, along with other parameters, into the filter. I don't know if the GPSDOs try to do a time series fit/model (at least for low order terms) to deal with things like diurnal variation. They could. I should note that the MCXO approach, popularized as a "better TCXO" in a small low power package, is also used in some software defined radios, except that there's no separate microcontroller. The frequency vs temperature characteristic is just embedded in the other algorithms in the radio's host processor.
CA
Chris Albertson
Tue, Aug 28, 2012 4:15 PM

On Mon, Aug 27, 2012 at 11:45 PM, WB6BNQ wb6bnq@cox.net wrote:

Chris,

That is a big no !  What the Thunderbolt is doing is adjusting the OCXO to
keep
it aligned with a known reference from the GPS system.  The fact that a
microprocessor is involved is from a totally different perspective.

Yes but but the t-bolt "remembers" the corrections is was applying and
continues to apply them whenthe GPS signal goes away.  I'm pretty sure one
of the things it remembers is the aging rate.

I know the "real" uP controlled XOs use programmed data for aging rate and
temp compensation the t-bolt gts it's own data by comparing with GPS.

Chris Albertson
Redondo Beach, California

On Mon, Aug 27, 2012 at 11:45 PM, WB6BNQ <wb6bnq@cox.net> wrote: > Chris, > > That is a big no ! What the Thunderbolt is doing is adjusting the OCXO to > keep > it aligned with a known reference from the GPS system. The fact that a > microprocessor is involved is from a totally different perspective. Yes but but the t-bolt "remembers" the corrections is was applying and continues to apply them whenthe GPS signal goes away. I'm pretty sure one of the things it remembers is the aging rate. I know the "real" uP controlled XOs use programmed data for aging rate and temp compensation the t-bolt gts it's own data by comparing with GPS. Chris Albertson Redondo Beach, California
CA
Chris Albertson
Tue, Aug 28, 2012 4:19 PM

On Tue, Aug 28, 2012 at 6:17 AM, Jim Lux jimlux@earthlink.net wrote:

The MCXO has a one time calibration for frequency vs temperature that's
programmed into it (and some use very clever ways to measure the
temperature).  The disciplining algorithms in a GPSDO are a bit smarter;

Yes, correct.  But I was talking about during hold-over.  t-bolt in
holdover is no longer a GPSDO.    It falls back on using a temperature and
aging model.

Chris Albertson
Redondo Beach, California

On Tue, Aug 28, 2012 at 6:17 AM, Jim Lux <jimlux@earthlink.net> wrote: > The MCXO has a one time calibration for frequency vs temperature that's > programmed into it (and some use *very* clever ways to measure the > temperature). The disciplining algorithms in a GPSDO are a bit smarter; > > Yes, correct. But I was talking about during hold-over. t-bolt in holdover is no longer a GPSDO. It falls back on using a temperature and aging model. Chris Albertson Redondo Beach, California
DL
Don Latham
Tue, Aug 28, 2012 4:21 PM

thanks,Azelio. Neubig kindly sent the location for the references.
Another good company, like the Wenzel site.
Don

Azelio Boriani

www.axtal.com/data/publ/aging_e.pdf

On Tue, Aug 28, 2012 at 2:32 AM, Don Latham djl@montana.com wrote:

Could you refresh me with the reference or email the paper? I'd like
to
give it a read.
Don

Azelio Boriani

My starting point is the paper by Neubig on the real vs. predicted
aging:
experimenting with the model and try to cut it around the Morion or

the

Oscilloquartz OCXOs.

On Tue, Aug 28, 2012 at 12:15 AM, Don Latham djl@montana.com

wrote:

There are ways of generating models, such as ARIMA, using past
behavior
and the models used to tweak filters and so on. The shifts would

have

to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for

a

given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO

taking

the

action from a suitable model should help in speeding up the

adjustments

and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com

wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a

suitable

level

for
that input. I have the 53132A and the 53181A so I can test

(with

the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of

a

model

that
can be general enough so that it can be "steered", adjusting

its

parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig

wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B

inputs

are

specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to

100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only

in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger

mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I

doubt

if

anyone can ;) Aging predictions can only be made on

individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited

time.

Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger

problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min

50MHz,

-20dBm
the problem is on A input only? Maybe 150MHz for the A input,

being

beyond
the official specifications, requires more signal on one

counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real

and

predicted
aging of crystal oscillators. I'm trying to find a suitable

model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a

Kalman

filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig

wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and

53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ

at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10

dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also

at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can

give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To

unsubscribe,

go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe,

go

to

https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and

argument

are as significant as experiment, for thence comes quiet to the

mind."

R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and

argument

are as significant as experiment, for thence comes quiet to the

mind."

R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
De Erroribus Medicorum, R. Bacon, 13th century.
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

thanks,Azelio. Neubig kindly sent the location for the references. Another good company, like the Wenzel site. Don Azelio Boriani > www.*axtal*.com/data/publ/*aging*_e.pdf > > > On Tue, Aug 28, 2012 at 2:32 AM, Don Latham <djl@montana.com> wrote: > >> Could you refresh me with the reference or email the paper? I'd like >> to >> give it a read. >> Don >> >> Azelio Boriani >> > My starting point is the paper by Neubig on the real vs. predicted >> > aging: >> > experimenting with the model and try to cut it around the Morion or >> the >> > Oscilloquartz OCXOs. >> > >> > On Tue, Aug 28, 2012 at 12:15 AM, Don Latham <djl@montana.com> >> wrote: >> > >> >> There are ways of generating models, such as ARIMA, using past >> >> behavior >> >> and the models used to tweak filters and so on. The shifts would >> have >> >> to >> >> be accounted for as they occur, seems to me. Now the interesting >> >> possibility is that a shift does not alter the underlying model for >> a >> >> given crystal???? >> >> Has to be tested. >> >> Don >> >> >> >> Azelio Boriani >> >> > So it can't be done... the possibility to discipline an OCXO >> taking >> >> the >> >> > action from a suitable model should help in speeding up the >> >> adjustments >> >> > and >> >> > avoiding the humps as seen in the Allan deviation plot. >> >> > >> >> > On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> >> wrote: >> >> > >> >> >> Isn't crystal aging marked by random jumps? perhaps lattice >> >> >> rearrangement? Really hard if not impossible to model... >> >> >> Don >> >> >> >> >> >> Azelio Boriani >> >> >> > 30mV are -13dBm so maybe it is why the -20dBm is not a >> suitable >> >> >> level >> >> >> > for >> >> >> > that input. I have the 53132A and the 53181A so I can test >> (with >> >> the >> >> >> R&S >> >> >> > SML01). Is this problem new? Never tested before? >> >> >> > >> >> >> > About the OCXO aging: OK, I'm interested in the development of >> a >> >> >> model >> >> >> > that >> >> >> > can be general enough so that it can be "steered", adjusting >> its >> >> >> > parameters, during the OCXO life. The data (to adjust the >> >> >> parameters) >> >> >> > will >> >> >> > be gathered by the continuous monitoring with the GPS. >> >> >> > >> >> >> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig >> >> <bneubig@t-online.de> >> >> >> > wrote: >> >> >> > >> >> >> >> Hi Azelio, >> >> >> >> >> >> >> >> Sorry, the counters are 53131A, not 53151A. The A and B >> inputs >> >> are >> >> >> >> specified >> >> >> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to >> 100 >> >> >> MHz, >> >> >> >> <30 mV >> >> >> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. >> >> >> >> The problem can be at either input alone. The problem is only >> in >> >> >> the >> >> >> >> range >> >> >> >> between 100 MHz and <200 MHz, observed in automatic trigger >> >> mode. >> >> >> In >> >> >> >> manual >> >> >> >> trigger mode the counter does not trigger at all in that freq >> >> >> range. >> >> >> >> >> >> >> >> I cannot tell you an aging model for a particular OCXO, and I >> >> doubt >> >> >> if >> >> >> >> anyone can ;) Aging predictions can only be made on >> individual >> >> >> OCXO, >> >> >> >> unit >> >> >> >> by >> >> >> >> unit, and the prediction is only valid over a rather limited >> >> time. >> >> >> >> Therefore >> >> >> >> you must continuously monitor and update your prediction. >> >> >> >> >> >> >> >> Best regards >> >> >> >> >> >> >> >> Bernd >> >> >> >> DK1AG >> >> >> >> >> >> >> >> -----Ursprüngliche Nachricht----- >> >> >> >> Von: time-nuts-bounces@febo.com >> >> [mailto:time-nuts-bounces@febo.com] >> >> >> Im >> >> >> >> Auftrag von Azelio Boriani >> >> >> >> Gesendet: Freitag, 24. August 2012 17:28 >> >> >> >> An: Discussion of precise time and frequency measurement >> >> >> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger >> >> problems >> >> >> >> >> >> >> >> From the datasheet: >> >> >> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min >> >> 50MHz, >> >> >> >> -20dBm >> >> >> >> the problem is on A input only? Maybe 150MHz for the A input, >> >> being >> >> >> >> beyond >> >> >> >> the official specifications, requires more signal on one >> counter >> >> >> than >> >> >> >> the >> >> >> >> other. In the past they all were fine? >> >> >> >> I have read your paper on the correlation between the real >> and >> >> >> >> predicted >> >> >> >> aging of crystal oscillators. I'm trying to find a suitable >> >> model >> >> >> for >> >> >> >> the >> >> >> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a >> >> Kalman >> >> >> >> filter. >> >> >> >> >> >> >> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig >> >> <BNeubig@t-online.de> >> >> >> >> wrote: >> >> >> >> >> >> >> >> > Hi all, >> >> >> >> > >> >> >> >> > I have several HP/Agilent frequency counters 53151A (and >> >> 53152A) >> >> >> in >> >> >> >> my >> >> >> >> lab. >> >> >> >> > Several of them meanwhile show trigger problems. >> >> >> >> > >> >> >> >> > I have tested the A and B channel between 5 MHz and 200 MHZ >> at >> >> >> three >> >> >> >> > input levels of -10 dBm, 0 dBm and +10 dBm. >> >> >> >> > >> >> >> >> > Those with trigger problems show erroneous numbers at -10 >> dBm >> >> >> input >> >> >> >> > level rather selectively, i.e. mostly at 100 MHz, some also >> at >> >> >> 150 >> >> >> >> > MHz, while the are o.k. >> >> >> >> > >> >> >> >> > Has anyone on this list experienced similar defects and can >> >> give >> >> >> a >> >> >> >> > recommendation how to fix/repair them? >> >> >> >> > >> >> >> >> > >> >> >> >> > >> >> >> >> > Best regards >> >> >> >> > >> >> >> >> > >> >> >> >> > >> >> >> >> > Bernd >> >> >> >> > >> >> >> >> > DK1AG >> >> >> >> > >> >> >> >> > www.axtal.com >> >> >> >> > >> >> >> >> > _______________________________________________ >> >> >> >> > time-nuts mailing list -- time-nuts@febo.com To >> unsubscribe, >> >> go >> >> >> to >> >> >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> >> > and follow the instructions there. >> >> >> >> > >> >> >> >> _______________________________________________ >> >> >> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, >> go >> >> to >> >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> >> and follow the instructions there. >> >> >> >> >> >> >> >> >> >> >> >> _______________________________________________ >> >> >> >> time-nuts mailing list -- time-nuts@febo.com >> >> >> >> To unsubscribe, go to >> >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> >> and follow the instructions there. >> >> >> >> >> >> >> > _______________________________________________ >> >> >> > time-nuts mailing list -- time-nuts@febo.com >> >> >> > To unsubscribe, go to >> >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> > and follow the instructions there. >> >> >> > >> >> >> > >> >> >> >> >> >> >> >> >> -- >> >> >> "Neither the voice of authority nor the weight of reason and >> >> argument >> >> >> are as significant as experiment, for thence comes quiet to the >> >> mind." >> >> >> R. Bacon >> >> >> "If you don't know what it is, don't poke it." >> >> >> Ghost in the Shell >> >> >> >> >> >> >> >> >> Dr. Don Latham AJ7LL >> >> >> Six Mile Systems LLP >> >> >> 17850 Six Mile Road >> >> >> POB 134 >> >> >> Huson, MT, 59846 >> >> >> VOX 406-626-4304 >> >> >> www.lightningforensics.com >> >> >> www.sixmilesystems.com >> >> >> >> >> >> >> >> >> >> >> >> _______________________________________________ >> >> >> time-nuts mailing list -- time-nuts@febo.com >> >> >> To unsubscribe, go to >> >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> and follow the instructions there. >> >> >> >> >> > _______________________________________________ >> >> > time-nuts mailing list -- time-nuts@febo.com >> >> > To unsubscribe, go to >> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> > and follow the instructions there. >> >> > >> >> > >> >> >> >> >> >> -- >> >> "Neither the voice of authority nor the weight of reason and >> argument >> >> are as significant as experiment, for thence comes quiet to the >> mind." >> >> R. Bacon >> >> "If you don't know what it is, don't poke it." >> >> Ghost in the Shell >> >> >> >> >> >> Dr. Don Latham AJ7LL >> >> Six Mile Systems LLP >> >> 17850 Six Mile Road >> >> POB 134 >> >> Huson, MT, 59846 >> >> VOX 406-626-4304 >> >> www.lightningforensics.com >> >> www.sixmilesystems.com >> >> >> >> >> >> >> >> _______________________________________________ >> >> time-nuts mailing list -- time-nuts@febo.com >> >> To unsubscribe, go to >> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> and follow the instructions there. >> >> >> > _______________________________________________ >> > time-nuts mailing list -- time-nuts@febo.com >> > To unsubscribe, go to >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> > and follow the instructions there. >> > >> > >> >> >> -- >> "Neither the voice of authority nor the weight of reason and argument >> are as significant as experiment, for thence comes quiet to the mind." >> R. Bacon >> "If you don't know what it is, don't poke it." >> Ghost in the Shell >> >> >> Dr. Don Latham AJ7LL >> Six Mile Systems LLP >> 17850 Six Mile Road >> POB 134 >> Huson, MT, 59846 >> VOX 406-626-4304 >> www.lightningforensics.com >> www.sixmilesystems.com >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." De Erroribus Medicorum, R. Bacon, 13th century. "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
BC
Bob Camp
Tue, Aug 28, 2012 4:37 PM

Hi

I don't know that there is any data showing the TBolt adjusting for aging
during holdover. In other words: showing the holdover DAC voltage changing
at a completely constant temperature.

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Chris Albertson
Sent: Tuesday, August 28, 2012 12:19 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] oscillators

On Tue, Aug 28, 2012 at 6:17 AM, Jim Lux jimlux@earthlink.net wrote:

The MCXO has a one time calibration for frequency vs temperature that's
programmed into it (and some use very clever ways to measure the
temperature).  The disciplining algorithms in a GPSDO are a bit smarter;

Yes, correct.  But I was talking about during hold-over.  t-bolt in
holdover is no longer a GPSDO.    It falls back on using a temperature and
aging model.

Chris Albertson
Redondo Beach, California


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I don't know that there is any data showing the TBolt adjusting for aging during holdover. In other words: showing the holdover DAC voltage changing at a completely constant temperature. Bob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Chris Albertson Sent: Tuesday, August 28, 2012 12:19 PM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] oscillators On Tue, Aug 28, 2012 at 6:17 AM, Jim Lux <jimlux@earthlink.net> wrote: > The MCXO has a one time calibration for frequency vs temperature that's > programmed into it (and some use *very* clever ways to measure the > temperature). The disciplining algorithms in a GPSDO are a bit smarter; > > Yes, correct. But I was talking about during hold-over. t-bolt in holdover is no longer a GPSDO. It falls back on using a temperature and aging model. Chris Albertson Redondo Beach, California _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
SH
steve heidmann
Tue, Aug 28, 2012 7:13 PM

The then Hughes Aircraft had a nice hybrid (microelectronic ) version of this
idea produced in Newport Beach. 
--- On Mon, 8/27/12, Rick Karlquist richard@karlquist.com wrote:

From: Rick Karlquist richard@karlquist.com
Subject: Re: [time-nuts] HP/Agilet counter 53131A (not 53151A) trigger problems
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Date: Monday, August 27, 2012, 4:15 PM

Several decades ago, the concept of the "smart clock" arose
at what was then HP.  The idea was as discussed here to
"characterize" past aging, "predict" future aging, and
then "correct" the aging.  The goal wasn't to turn a quartz
oscillator into an atomic clock replacement, but simply
to get the oscillator through a 1 hour or so "holdover" time
during GPS outages.  It sort of worked for that very limited
purpose, but in general, past performance of HP crystals wasn't
a very good predictor of future results.  Crystals would age
in one direction for a while and possibly slow down as time
when on, but then then might start aging in the other direction.
There were also frequency jumps that were substantial and totally
random.  The reason why the HP crystals were unpredictable was
that all the deterministic processes such as mass preferentially
depositing on the crystal, so as to make the frequency age
lower, had been eliminated by years of manufacturing improvements.
The remaining processes were of the nature of quartz stress
relaxation that were very random.

Rick Karlquist

Don Latham wrote:

There are ways of generating models, such as ARIMA, using past behavior
and the models used to tweak filters and so on. The shifts would have to
be accounted for as they occur, seems to me. Now the interesting
possibility is that a shift does not alter the underlying model for a
given crystal????
Has to be tested.
Don

Azelio Boriani

So it can't be done... the possibility to discipline an OCXO taking the
action from a suitable model should help in speeding up the adjustments
and
avoiding the humps as seen in the Allan deviation plot.

On Mon, Aug 27, 2012 at 9:31 PM, Don Latham djl@montana.com wrote:

Isn't crystal aging marked by random jumps? perhaps lattice
rearrangement? Really hard if not impossible to model...
Don

Azelio Boriani

30mV are -13dBm so maybe it is why the -20dBm is not a suitable

level

for
that input. I have the 53132A and the 53181A so I can test (with the

R&S

SML01). Is this problem new? Never tested before?

About the OCXO aging: OK, I'm interested in the development of a

model

that
can be general enough so that it can be "steered", adjusting its
parameters, during the OCXO life. The data (to adjust the

parameters)

will
be gathered by the continuous monitoring with the GPS.

On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig bneubig@t-online.de
wrote:

Hi Azelio,

Sorry, the counters are 53131A, not 53151A. The A and B inputs are
specified
from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100

MHz,

<30 mV
(rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz.
The problem can be at either input alone. The problem is only in

the

range
between 100 MHz and <200 MHz, observed in automatic trigger mode.

In

manual
trigger mode the counter does not trigger at all in that freq

range.

I cannot tell you an aging model for a particular OCXO, and I doubt

if

anyone can ;) Aging predictions can only be made on individual

OCXO,

unit
by
unit, and the prediction is only valid over a rather limited time.
Therefore
you must continuously monitor and update your prediction.

Best regards

Bernd
DK1AG

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com]

Im

Auftrag von Azelio Boriani
Gesendet: Freitag, 24. August 2012 17:28
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems

From the datasheet:
A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz,
-20dBm
the problem is on A input only? Maybe 150MHz for the A input, being
beyond
the official specifications, requires more signal on one counter

than

the
other. In the past they all were fine?
I have read your paper on the correlation between the real and
predicted
aging of crystal oscillators. I'm trying to find a suitable model

for

the
Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman
filter.

On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig BNeubig@t-online.de
wrote:

Hi all,

I have several HP/Agilent frequency counters 53151A (and 53152A)

in

my
lab.

Several of them meanwhile show trigger problems.

I have tested the A and B channel between 5 MHz and 200 MHZ at

three

input levels of -10 dBm, 0 dBm and +10 dBm.

Those with trigger problems show erroneous numbers at -10 dBm

input

level rather selectively, i.e. mostly at 100 MHz, some also at

150

MHz, while the are o.k.

Has anyone on this list experienced similar defects and can give

a

recommendation how to fix/repair them?

Best regards

Bernd

DK1AG

www.axtal.com


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go

to


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
R. Bacon
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

The then Hughes Aircraft had a nice hybrid (microelectronic ) version of this idea produced in Newport Beach.  --- On Mon, 8/27/12, Rick Karlquist <richard@karlquist.com> wrote: From: Rick Karlquist <richard@karlquist.com> Subject: Re: [time-nuts] HP/Agilet counter 53131A (not 53151A) trigger problems To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Date: Monday, August 27, 2012, 4:15 PM Several decades ago, the concept of the "smart clock" arose at what was then HP.  The idea was as discussed here to "characterize" past aging, "predict" future aging, and then "correct" the aging.  The goal wasn't to turn a quartz oscillator into an atomic clock replacement, but simply to get the oscillator through a 1 hour or so "holdover" time during GPS outages.  It sort of worked for that very limited purpose, but in general, past performance of HP crystals wasn't a very good predictor of future results.  Crystals would age in one direction for a while and possibly slow down as time when on, but then then might start aging in the other direction. There were also frequency jumps that were substantial and totally random.  The reason why the HP crystals were unpredictable was that all the deterministic processes such as mass preferentially depositing on the crystal, so as to make the frequency age lower, had been eliminated by years of manufacturing improvements. The remaining processes were of the nature of quartz stress relaxation that were very random. Rick Karlquist Don Latham wrote: > There are ways of generating models, such as ARIMA, using past behavior > and the models used to tweak filters and so on. The shifts would have to > be accounted for as they occur, seems to me. Now the interesting > possibility is that a shift does not alter the underlying model for a > given crystal???? > Has to be tested. > Don > > Azelio Boriani >> So it can't be done... the possibility to discipline an OCXO taking the >> action from a suitable model should help in speeding up the adjustments >> and >> avoiding the humps as seen in the Allan deviation plot. >> >> On Mon, Aug 27, 2012 at 9:31 PM, Don Latham <djl@montana.com> wrote: >> >>> Isn't crystal aging marked by random jumps? perhaps lattice >>> rearrangement? Really hard if not impossible to model... >>> Don >>> >>> Azelio Boriani >>> > 30mV are -13dBm so maybe it is why the -20dBm is not a suitable >>> level >>> > for >>> > that input. I have the 53132A and the 53181A so I can test (with the >>> R&S >>> > SML01). Is this problem new? Never tested before? >>> > >>> > About the OCXO aging: OK, I'm interested in the development of a >>> model >>> > that >>> > can be general enough so that it can be "steered", adjusting its >>> > parameters, during the OCXO life. The data (to adjust the >>> parameters) >>> > will >>> > be gathered by the continuous monitoring with the GPS. >>> > >>> > On Mon, Aug 27, 2012 at 7:01 PM, Bernd Neubig <bneubig@t-online.de> >>> > wrote: >>> > >>> >> Hi Azelio, >>> >> >>> >> Sorry, the counters are 53131A, not 53151A. The A and B inputs are >>> >> specified >>> >> from DC to 225 MHz with a sensitivity of y20 mV (rms) up to 100 >>> MHz, >>> >> <30 mV >>> >> (rms) 100 ~ 200 MHz, and <50 mV (rms) 200 ~ 225 MHz. >>> >> The problem can be at either input alone. The problem is only in >>> the >>> >> range >>> >> between 100 MHz and <200 MHz, observed in automatic trigger mode. >>> In >>> >> manual >>> >> trigger mode the counter does not trigger at all in that freq >>> range. >>> >> >>> >> I cannot tell you an aging model for a particular OCXO, and I doubt >>> if >>> >> anyone can ;) Aging predictions can only be made on individual >>> OCXO, >>> >> unit >>> >> by >>> >> unit, and the prediction is only valid over a rather limited time. >>> >> Therefore >>> >> you must continuously monitor and update your prediction. >>> >> >>> >> Best regards >>> >> >>> >> Bernd >>> >> DK1AG >>> >> >>> >> -----Ursprüngliche Nachricht----- >>> >> Von: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] >>> Im >>> >> Auftrag von Azelio Boriani >>> >> Gesendet: Freitag, 24. August 2012 17:28 >>> >> An: Discussion of precise time and frequency measurement >>> >> Betreff: Re: [time-nuts] HP/Agilet counter 53151A trigger problems >>> >> >>> >> From the datasheet: >>> >> A input 10Hz..125MHz min level 25mVrms (-19dBm) B input min 50MHz, >>> >> -20dBm >>> >> the problem is on A input only? Maybe 150MHz for the A input, being >>> >> beyond >>> >> the official specifications, requires more signal on one counter >>> than >>> >> the >>> >> other. In the past they all were fine? >>> >> I have read your paper on the correlation between the real and >>> >> predicted >>> >> aging of crystal oscillators. I'm trying to find a suitable model >>> for >>> >> the >>> >> Morion MV201 OCXO or the Oscilloquartz 8663 to implement a Kalman >>> >> filter. >>> >> >>> >> On Fri, Aug 24, 2012 at 3:34 PM, Bernd Neubig <BNeubig@t-online.de> >>> >> wrote: >>> >> >>> >> > Hi all, >>> >> > >>> >> > I have several HP/Agilent frequency counters 53151A (and 53152A) >>> in >>> >> my >>> >> lab. >>> >> > Several of them meanwhile show trigger problems. >>> >> > >>> >> > I have tested the A and B channel between 5 MHz and 200 MHZ at >>> three >>> >> > input levels of -10 dBm, 0 dBm and +10 dBm. >>> >> > >>> >> > Those with trigger problems show erroneous numbers at -10 dBm >>> input >>> >> > level rather selectively, i.e. mostly at 100 MHz, some also at >>> 150 >>> >> > MHz, while the are o.k. >>> >> > >>> >> > Has anyone on this list experienced similar defects and can give >>> a >>> >> > recommendation how to fix/repair them? >>> >> > >>> >> > >>> >> > >>> >> > Best regards >>> >> > >>> >> > >>> >> > >>> >> > Bernd >>> >> > >>> >> > DK1AG >>> >> > >>> >> > www.axtal.com >>> >> > >>> >> > _______________________________________________ >>> >> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go >>> to >>> >> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >> > and follow the instructions there. >>> >> > >>> >> _______________________________________________ >>> >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >> and follow the instructions there. >>> >> >>> >> >>> >> _______________________________________________ >>> >> time-nuts mailing list -- time-nuts@febo.com >>> >> To unsubscribe, go to >>> >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >> and follow the instructions there. >>> >> >>> > _______________________________________________ >>> > time-nuts mailing list -- time-nuts@febo.com >>> > To unsubscribe, go to >>> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> > and follow the instructions there. >>> > >>> > >>> >>> >>> -- >>> "Neither the voice of authority nor the weight of reason and argument >>> are as significant as experiment, for thence comes quiet to the mind." >>> R. Bacon >>> "If you don't know what it is, don't poke it." >>> Ghost in the Shell >>> >>> >>> Dr. Don Latham AJ7LL >>> Six Mile Systems LLP >>> 17850 Six Mile Road >>> POB 134 >>> Huson, MT, 59846 >>> VOX 406-626-4304 >>> www.lightningforensics.com >>> www.sixmilesystems.com >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> > > > -- > "Neither the voice of authority nor the weight of reason and argument > are as significant as experiment, for thence comes quiet to the mind." > R. Bacon > "If you don't know what it is, don't poke it." > Ghost in the Shell > > > Dr. Don Latham AJ7LL > Six Mile Systems LLP > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > www.lightningforensics.com > www.sixmilesystems.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
R(
Richard (Rick) Karlquist
Thu, Aug 30, 2012 3:45 AM

On 8/27/2012 11:45 PM, WB6BNQ wrote:

A microprocessor controlled XO is a non oven crystal oscillator system that has
additional computational control providing a bit more than just mere passive
temperature compensation.  The additional computational capability deals with
having coefficients of that particular oscillator's behavior pre coded to
compensate for the nonlinear behavior over a given temperature rang

It doesn't use coefficients.  It has a look up table of frequency vs
temperature.

A microprocessor controlled XO system allows for using cheap crystals with
minimum processing time and costs.  Because of limited storage space there is no

No it doesn't use a cheap crystal.  It uses a special SC cut crystal.
This crystal could very easily cost more than an OCXO crystal.

way for the system to have enough data to even try to compete with the quality of
a decent OCXO.  Beyond its initial calibration setup, it has no way of keeping it
tied to a known reference, like the Thunderbolt is doing.

Bill....WB6BNQ

An MCXO is a very good, but expensive TCXO.  Only temperature, not aging
is corrected.

It has nothing to do with "smart clocks".

Rick Karlquist N6RK

On 8/27/2012 11:45 PM, WB6BNQ wrote: > > > A microprocessor controlled XO is a non oven crystal oscillator system that has > additional computational control providing a bit more than just mere passive > temperature compensation. The additional computational capability deals with > having coefficients of that particular oscillator's behavior pre coded to > compensate for the nonlinear behavior over a given temperature rang It doesn't use coefficients. It has a look up table of frequency vs temperature. > > A microprocessor controlled XO system allows for using cheap crystals with > minimum processing time and costs. Because of limited storage space there is no No it doesn't use a cheap crystal. It uses a *special* SC cut crystal. This crystal could very easily cost more than an OCXO crystal. > way for the system to have enough data to even try to compete with the quality of > a decent OCXO. Beyond its initial calibration setup, it has no way of keeping it > tied to a known reference, like the Thunderbolt is doing. > > Bill....WB6BNQ An MCXO is a very good, but expensive TCXO. Only temperature, not aging is corrected. It has nothing to do with "smart clocks". Rick Karlquist N6RK
NM
Neville Michie
Thu, Aug 30, 2012 4:12 AM

Does anyone know about what technology is used in Swiss watches to get much better performance from
their xtals than you might expect?
I assumed that they had look up lists to insert extra counts to compensate for ambient variations,
but I have never heard any details.
cheers,
Neville Michie

On 30/08/2012, at 1:45 PM, Richard (Rick) Karlquist wrote:

On 8/27/2012 11:45 PM, WB6BNQ wrote:

A microprocessor controlled XO is a non oven crystal oscillator system that has
additional computational control providing a bit more than just mere passive
temperature compensation.  The additional computational capability deals with
having coefficients of that particular oscillator's behavior pre coded to
compensate for the nonlinear behavior over a given temperature rang

It doesn't use coefficients.  It has a look up table of frequency vs
temperature.

A microprocessor controlled XO system allows for using cheap crystals with
minimum processing time and costs.  Because of limited storage space there is no

No it doesn't use a cheap crystal.  It uses a special SC cut crystal.
This crystal could very easily cost more than an OCXO crystal.

way for the system to have enough data to even try to compete with the quality of
a decent OCXO.  Beyond its initial calibration setup, it has no way of keeping it
tied to a known reference, like the Thunderbolt is doing.

Bill....WB6BNQ

An MCXO is a very good, but expensive TCXO.  Only temperature, not aging is corrected.

It has nothing to do with "smart clocks".

Rick Karlquist N6RK


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Does anyone know about what technology is used in Swiss watches to get much better performance from their xtals than you might expect? I assumed that they had look up lists to insert extra counts to compensate for ambient variations, but I have never heard any details. cheers, Neville Michie On 30/08/2012, at 1:45 PM, Richard (Rick) Karlquist wrote: > > On 8/27/2012 11:45 PM, WB6BNQ wrote: >> >> >> A microprocessor controlled XO is a non oven crystal oscillator system that has >> additional computational control providing a bit more than just mere passive >> temperature compensation. The additional computational capability deals with >> having coefficients of that particular oscillator's behavior pre coded to >> compensate for the nonlinear behavior over a given temperature rang > > It doesn't use coefficients. It has a look up table of frequency vs > temperature. > >> >> A microprocessor controlled XO system allows for using cheap crystals with >> minimum processing time and costs. Because of limited storage space there is no > > No it doesn't use a cheap crystal. It uses a *special* SC cut crystal. > This crystal could very easily cost more than an OCXO crystal. > >> way for the system to have enough data to even try to compete with the quality of >> a decent OCXO. Beyond its initial calibration setup, it has no way of keeping it >> tied to a known reference, like the Thunderbolt is doing. >> >> Bill....WB6BNQ > > An MCXO is a very good, but expensive TCXO. Only temperature, not aging is corrected. > > It has nothing to do with "smart clocks". > > Rick Karlquist N6RK > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MC
mike cook
Thu, Aug 30, 2012 4:18 AM

nippedLe 30/08/2012 06:12, Neville Michie a écrit :

Does anyone know about what technology is used in Swiss watches to get much better performance from
their xtals than you might expect?
I assumed that they had look up lists to insert extra counts to compensate for ambient variations,
but I have never heard any details.
cheers,
Neville Michie

check out
http://forums.watchuseek.com/f9/thermocompensation-methods-movements-2087.html
<snipped>

--
Les chiens aboient, et la caravane passe.

nippedLe 30/08/2012 06:12, Neville Michie a écrit : > Does anyone know about what technology is used in Swiss watches to get much better performance from > their xtals than you might expect? > I assumed that they had look up lists to insert extra counts to compensate for ambient variations, > but I have never heard any details. > cheers, > Neville Michie > > check out <http://forums.watchuseek.com/f9/thermocompensation-methods-movements-2087.html> <snipped> -- Les chiens aboient, et la caravane passe.
CH
Chuck Harris
Thu, Aug 30, 2012 4:21 AM

Well, basically the temperature of a wrist watch is very
constant at around 90F.  That and a table lookup that gives
an adjustment for number of cycles per tick vs temperature.

All quartz watches since about 1990 are microprocessor based,
and have a table lookup.  They can only be regulated using
a special computer interface provided by the manufacturer.

-Chuck Harris

Neville Michie wrote:

Does anyone know about what technology is used in Swiss watches to get much better performance from
their xtals than you might expect?
I assumed that they had look up lists to insert extra counts to compensate for ambient variations,
but I have never heard any details.
cheers,
Neville Michie

On 30/08/2012, at 1:45 PM, Richard (Rick) Karlquist wrote:

On 8/27/2012 11:45 PM, WB6BNQ wrote:

A microprocessor controlled XO is a non oven crystal oscillator system that has
additional computational control providing a bit more than just mere passive
temperature compensation.  The additional computational capability deals with
having coefficients of that particular oscillator's behavior pre coded to
compensate for the nonlinear behavior over a given temperature rang

It doesn't use coefficients.  It has a look up table of frequency vs
temperature.

A microprocessor controlled XO system allows for using cheap crystals with
minimum processing time and costs.  Because of limited storage space there is no

No it doesn't use a cheap crystal.  It uses a special SC cut crystal.
This crystal could very easily cost more than an OCXO crystal.

way for the system to have enough data to even try to compete with the quality of
a decent OCXO.  Beyond its initial calibration setup, it has no way of keeping it
tied to a known reference, like the Thunderbolt is doing.

Bill....WB6BNQ

An MCXO is a very good, but expensive TCXO.  Only temperature, not aging is corrected.

It has nothing to do with "smart clocks".

Rick Karlquist N6RK


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Well, basically the temperature of a wrist watch is very constant at around 90F. That and a table lookup that gives an adjustment for number of cycles per tick vs temperature. All quartz watches since about 1990 are microprocessor based, and have a table lookup. They can only be regulated using a special computer interface provided by the manufacturer. -Chuck Harris Neville Michie wrote: > Does anyone know about what technology is used in Swiss watches to get much better performance from > their xtals than you might expect? > I assumed that they had look up lists to insert extra counts to compensate for ambient variations, > but I have never heard any details. > cheers, > Neville Michie > > > > > > > On 30/08/2012, at 1:45 PM, Richard (Rick) Karlquist wrote: > >> >> On 8/27/2012 11:45 PM, WB6BNQ wrote: >>> >>> >>> A microprocessor controlled XO is a non oven crystal oscillator system that has >>> additional computational control providing a bit more than just mere passive >>> temperature compensation. The additional computational capability deals with >>> having coefficients of that particular oscillator's behavior pre coded to >>> compensate for the nonlinear behavior over a given temperature rang >> >> It doesn't use coefficients. It has a look up table of frequency vs >> temperature. >> >>> >>> A microprocessor controlled XO system allows for using cheap crystals with >>> minimum processing time and costs. Because of limited storage space there is no >> >> No it doesn't use a cheap crystal. It uses a *special* SC cut crystal. >> This crystal could very easily cost more than an OCXO crystal. >> >>> way for the system to have enough data to even try to compete with the quality of >>> a decent OCXO. Beyond its initial calibration setup, it has no way of keeping it >>> tied to a known reference, like the Thunderbolt is doing. >>> >>> Bill....WB6BNQ >> >> An MCXO is a very good, but expensive TCXO. Only temperature, not aging is corrected. >> >> It has nothing to do with "smart clocks". >> >> Rick Karlquist N6RK >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
JL
Jim Lux
Thu, Aug 30, 2012 12:54 PM

On 8/29/12 8:45 PM, Richard (Rick) Karlquist wrote:

On 8/27/2012 11:45 PM, WB6BNQ wrote:

A microprocessor controlled XO is a non oven crystal oscillator system
that has
additional computational control providing a bit more than just mere
passive
temperature compensation.  The additional computational capability
deals with
having coefficients of that particular oscillator's behavior pre coded to
compensate for the nonlinear behavior over a given temperature rang

It doesn't use coefficients.  It has a look up table of frequency vs
temperature.

A microprocessor controlled XO system allows for using cheap crystals
with
minimum processing time and costs.  Because of limited storage space
there is no

No it doesn't use a cheap crystal.  It uses a special SC cut crystal.
This crystal could very easily cost more than an OCXO crystal.

Isn't the crystal cut (and the circuit designed) in such a way that it
supports both the fundamental and the third overtone simultaneously, and
comparing the frequencies of the two (or, more accurately, fF- fthird/3)
is how they measure the temperature, which is then used to look up the
correction factor in a PROM.

On 8/29/12 8:45 PM, Richard (Rick) Karlquist wrote: > > On 8/27/2012 11:45 PM, WB6BNQ wrote: >> >> >> A microprocessor controlled XO is a non oven crystal oscillator system >> that has >> additional computational control providing a bit more than just mere >> passive >> temperature compensation. The additional computational capability >> deals with >> having coefficients of that particular oscillator's behavior pre coded to >> compensate for the nonlinear behavior over a given temperature rang > > It doesn't use coefficients. It has a look up table of frequency vs > temperature. > >> >> A microprocessor controlled XO system allows for using cheap crystals >> with >> minimum processing time and costs. Because of limited storage space >> there is no > > No it doesn't use a cheap crystal. It uses a *special* SC cut crystal. > This crystal could very easily cost more than an OCXO crystal. > Isn't the crystal cut (and the circuit designed) in such a way that it supports both the fundamental and the third overtone simultaneously, and comparing the frequencies of the two (or, more accurately, fF- fthird/3) is how they measure the temperature, which is then used to look up the correction factor in a PROM.
BC
Bob Camp
Thu, Aug 30, 2012 4:33 PM

Hi

The fundamental / third approach is one of several possible ways to go. You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a good
thermometer. Some have even gone as far as to mount a thermistor on the
crystal.

No matter what you do, it adds cost.

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Jim Lux
Sent: Thursday, August 30, 2012 8:55 AM
To: time-nuts@febo.com
Subject: Re: [time-nuts] oscillators

On 8/29/12 8:45 PM, Richard (Rick) Karlquist wrote:

On 8/27/2012 11:45 PM, WB6BNQ wrote:

A microprocessor controlled XO is a non oven crystal oscillator system
that has
additional computational control providing a bit more than just mere
passive
temperature compensation.  The additional computational capability
deals with
having coefficients of that particular oscillator's behavior pre coded to
compensate for the nonlinear behavior over a given temperature rang

It doesn't use coefficients.  It has a look up table of frequency vs
temperature.

A microprocessor controlled XO system allows for using cheap crystals
with
minimum processing time and costs.  Because of limited storage space
there is no

No it doesn't use a cheap crystal.  It uses a special SC cut crystal.
This crystal could very easily cost more than an OCXO crystal.

Isn't the crystal cut (and the circuit designed) in such a way that it
supports both the fundamental and the third overtone simultaneously, and
comparing the frequencies of the two (or, more accurately, fF- fthird/3)
is how they measure the temperature, which is then used to look up the
correction factor in a PROM.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The fundamental / third approach is one of several possible ways to go. You can also run an SC on the B and C modes to get thermometer data. Early implementations used a pair of independent blanks, one cut to be a good thermometer. Some have even gone as far as to mount a thermistor on the crystal. No matter what you do, it adds cost. Bob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Jim Lux Sent: Thursday, August 30, 2012 8:55 AM To: time-nuts@febo.com Subject: Re: [time-nuts] oscillators On 8/29/12 8:45 PM, Richard (Rick) Karlquist wrote: > > On 8/27/2012 11:45 PM, WB6BNQ wrote: >> >> >> A microprocessor controlled XO is a non oven crystal oscillator system >> that has >> additional computational control providing a bit more than just mere >> passive >> temperature compensation. The additional computational capability >> deals with >> having coefficients of that particular oscillator's behavior pre coded to >> compensate for the nonlinear behavior over a given temperature rang > > It doesn't use coefficients. It has a look up table of frequency vs > temperature. > >> >> A microprocessor controlled XO system allows for using cheap crystals >> with >> minimum processing time and costs. Because of limited storage space >> there is no > > No it doesn't use a cheap crystal. It uses a *special* SC cut crystal. > This crystal could very easily cost more than an OCXO crystal. > Isn't the crystal cut (and the circuit designed) in such a way that it supports both the fundamental and the third overtone simultaneously, and comparing the frequencies of the two (or, more accurately, fF- fthird/3) is how they measure the temperature, which is then used to look up the correction factor in a PROM. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
MD
Magnus Danielson
Thu, Aug 30, 2012 8:30 PM

On 08/30/2012 06:33 PM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to go. You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a good
thermometer. Some have even gone as far as to mount a thermistor on the
crystal.

No matter what you do, it adds cost.

The reason for doing it is to lower powerconsumption in critical
applications.

Cheers,
Magnus

On 08/30/2012 06:33 PM, Bob Camp wrote: > Hi > > The fundamental / third approach is one of several possible ways to go. You > can also run an SC on the B and C modes to get thermometer data. Early > implementations used a pair of independent blanks, one cut to be a good > thermometer. Some have even gone as far as to mount a thermistor on the > crystal. > > No matter what you do, it adds cost. The reason for doing it is to lower powerconsumption in critical applications. Cheers, Magnus
AB
Azelio Boriani
Thu, Aug 30, 2012 8:37 PM

Interesting this third overtone/fundamental3 idea but what makes the third
overtone not equal to the fundamental
3? Is the crystal operated in series
or parallel resonance?

On Thu, Aug 30, 2012 at 10:30 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:

On 08/30/2012 06:33 PM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to go.
You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a good
thermometer. Some have even gone as far as to mount a thermistor on the
crystal.

No matter what you do, it adds cost.

The reason for doing it is to lower powerconsumption in critical
applications.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Interesting this third overtone/fundamental*3 idea but what makes the third overtone not equal to the fundamental*3? Is the crystal operated in series or parallel resonance? On Thu, Aug 30, 2012 at 10:30 PM, Magnus Danielson < magnus@rubidium.dyndns.org> wrote: > On 08/30/2012 06:33 PM, Bob Camp wrote: > >> Hi >> >> The fundamental / third approach is one of several possible ways to go. >> You >> can also run an SC on the B and C modes to get thermometer data. Early >> implementations used a pair of independent blanks, one cut to be a good >> thermometer. Some have even gone as far as to mount a thermistor on the >> crystal. >> >> No matter what you do, it adds cost. >> > > The reason for doing it is to lower powerconsumption in critical > applications. > > Cheers, > Magnus > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
RK
Rick Karlquist
Thu, Aug 30, 2012 9:10 PM

Azelio Boriani wrote:

Interesting this third overtone/fundamental3 idea but what makes the
third
overtone not equal to the fundamental
3? Is the crystal operated in series
or parallel resonance?

That is where the special SC cut crystal comes into play.
It's not so much that the 3rd OT is on a different frequency, but
rather that it has a different and known tempco relative to the
fundamental, so it can be used as a thermometer.  It would be
interesting to take a crystal out of a random 10811 and compare the
frequencies of the fundamental and 3rd OT vs temperature to see
if that crystal would accidentally make a good thermometer.

Rick Karlquist N6RK

Azelio Boriani wrote: > Interesting this third overtone/fundamental*3 idea but what makes the > third > overtone not equal to the fundamental*3? Is the crystal operated in series > or parallel resonance? > That is where the special SC cut crystal comes into play. It's not so much that the 3rd OT is on a different frequency, but rather that it has a different and known tempco relative to the fundamental, so it can be used as a thermometer. It would be interesting to take a crystal out of a random 10811 and compare the frequencies of the fundamental and 3rd OT vs temperature to see if that crystal would accidentally make a good thermometer. Rick Karlquist N6RK
BC
Bob Camp
Thu, Aug 30, 2012 9:13 PM

Hi

It turns out that the fundamental is "not quite" 1/3 of the third overtone. You can impact the degree of offset by changing the contour of the blank. The temperature coefficient is also different on the fundamental. The net result is that you can get a pretty good thermometer reading by mixing one with the other divided by 3.

Bob

On Aug 30, 2012, at 4:37 PM, Azelio Boriani azelio.boriani@screen.it wrote:

Interesting this third overtone/fundamental3 idea but what makes the third
overtone not equal to the fundamental
3? Is the crystal operated in series
or parallel resonance?

On Thu, Aug 30, 2012 at 10:30 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:

On 08/30/2012 06:33 PM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to go.
You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a good
thermometer. Some have even gone as far as to mount a thermistor on the
crystal.

No matter what you do, it adds cost.

The reason for doing it is to lower powerconsumption in critical
applications.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi It turns out that the fundamental is "not quite" 1/3 of the third overtone. You can impact the degree of offset by changing the contour of the blank. The temperature coefficient is also different on the fundamental. The net result is that you can get a pretty good thermometer reading by mixing one with the other divided by 3. Bob On Aug 30, 2012, at 4:37 PM, Azelio Boriani <azelio.boriani@screen.it> wrote: > Interesting this third overtone/fundamental*3 idea but what makes the third > overtone not equal to the fundamental*3? Is the crystal operated in series > or parallel resonance? > > On Thu, Aug 30, 2012 at 10:30 PM, Magnus Danielson < > magnus@rubidium.dyndns.org> wrote: > >> On 08/30/2012 06:33 PM, Bob Camp wrote: >> >>> Hi >>> >>> The fundamental / third approach is one of several possible ways to go. >>> You >>> can also run an SC on the B and C modes to get thermometer data. Early >>> implementations used a pair of independent blanks, one cut to be a good >>> thermometer. Some have even gone as far as to mount a thermistor on the >>> crystal. >>> >>> No matter what you do, it adds cost. >>> >> >> The reason for doing it is to lower powerconsumption in critical >> applications. >> >> Cheers, >> Magnus >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Thu, Aug 30, 2012 9:15 PM

Hi

You can do the same thing with an AT as with an SC. They both exhibit the fundamental / third offset and tempo delta.

Bob

On Aug 30, 2012, at 5:10 PM, "Rick Karlquist" richard@karlquist.com wrote:

Azelio Boriani wrote:

Interesting this third overtone/fundamental3 idea but what makes the
third
overtone not equal to the fundamental
3? Is the crystal operated in series
or parallel resonance?

That is where the special SC cut crystal comes into play.
It's not so much that the 3rd OT is on a different frequency, but
rather that it has a different and known tempco relative to the
fundamental, so it can be used as a thermometer.  It would be
interesting to take a crystal out of a random 10811 and compare the
frequencies of the fundamental and 3rd OT vs temperature to see
if that crystal would accidentally make a good thermometer.

Rick Karlquist N6RK


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi You can do the same thing with an AT as with an SC. They both exhibit the fundamental / third offset and tempo delta. Bob On Aug 30, 2012, at 5:10 PM, "Rick Karlquist" <richard@karlquist.com> wrote: > Azelio Boriani wrote: >> Interesting this third overtone/fundamental*3 idea but what makes the >> third >> overtone not equal to the fundamental*3? Is the crystal operated in series >> or parallel resonance? >> > > That is where the special SC cut crystal comes into play. > It's not so much that the 3rd OT is on a different frequency, but > rather that it has a different and known tempco relative to the > fundamental, so it can be used as a thermometer. It would be > interesting to take a crystal out of a random 10811 and compare the > frequencies of the fundamental and 3rd OT vs temperature to see > if that crystal would accidentally make a good thermometer. > > Rick Karlquist N6RK > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Thu, Aug 30, 2012 9:23 PM

On 08/30/2012 11:10 PM, Rick Karlquist wrote:

Azelio Boriani wrote:

Interesting this third overtone/fundamental3 idea but what makes the
third
overtone not equal to the fundamental
3? Is the crystal operated in series
or parallel resonance?

That is where the special SC cut crystal comes into play.
It's not so much that the 3rd OT is on a different frequency, but
rather that it has a different and known tempco relative to the
fundamental, so it can be used as a thermometer.  It would be
interesting to take a crystal out of a random 10811 and compare the
frequencies of the fundamental and 3rd OT vs temperature to see
if that crystal would accidentally make a good thermometer.

All the different modes have different temperature dependencies. Being
able to maintain dual oscillation on both over the range seems like a
reasonable challenge.

Cheers,
Magnus

On 08/30/2012 11:10 PM, Rick Karlquist wrote: > Azelio Boriani wrote: >> Interesting this third overtone/fundamental*3 idea but what makes the >> third >> overtone not equal to the fundamental*3? Is the crystal operated in series >> or parallel resonance? >> > > That is where the special SC cut crystal comes into play. > It's not so much that the 3rd OT is on a different frequency, but > rather that it has a different and known tempco relative to the > fundamental, so it can be used as a thermometer. It would be > interesting to take a crystal out of a random 10811 and compare the > frequencies of the fundamental and 3rd OT vs temperature to see > if that crystal would accidentally make a good thermometer. All the different modes have different temperature dependencies. Being able to maintain dual oscillation on both over the range seems like a reasonable challenge. Cheers, Magnus
JL
Jim Lux
Fri, Aug 31, 2012 12:01 AM

On 8/30/12 9:33 AM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to go. You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a good
thermometer. Some have even gone as far as to mount a thermistor on the
crystal.

I've used the "sensor on oscillator can" technique very recently on a
software defined radio.  Over the next year, I'll get some data on how
well it works in use.  The idea is to use a low power TCXO and a sensor,
rather than an OCXO, to meet a fairly tight frequency requirement (a few
hundred Hz out of 2 GHz)  The TCXO can get you do a few ppm or better.
The sensor should let me get about an order of magnitude better, and I
need to measure the temperature anyway, and it's easy to integrate with
the software waveform code.

On 8/30/12 9:33 AM, Bob Camp wrote: > Hi > > The fundamental / third approach is one of several possible ways to go. You > can also run an SC on the B and C modes to get thermometer data. Early > implementations used a pair of independent blanks, one cut to be a good > thermometer. Some have even gone as far as to mount a thermistor on the > crystal. > I've used the "sensor on oscillator can" technique very recently on a software defined radio. Over the next year, I'll get some data on how well it works in use. The idea is to use a low power TCXO and a sensor, rather than an OCXO, to meet a fairly tight frequency requirement (a few hundred Hz out of 2 GHz) The TCXO can get you do a few ppm or better. The sensor should let me get about an order of magnitude better, and I need to measure the temperature anyway, and it's easy to integrate with the software waveform code.
DL
Don Latham
Fri, Aug 31, 2012 12:47 AM

Elecraft seems to be having some trouble with the KX3 stability for WSJT
on 6m. They also have a temp sensor on the bard, apparently and are also
trying this out. I don't know any more about it than that.
Don

Jim Lux

On 8/30/12 9:33 AM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to
go. You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a
good
thermometer. Some have even gone as far as to mount a thermistor on
the
crystal.

I've used the "sensor on oscillator can" technique very recently on a
software defined radio.  Over the next year, I'll get some data on how
well it works in use.  The idea is to use a low power TCXO and a sensor,
rather than an OCXO, to meet a fairly tight frequency requirement (a few
hundred Hz out of 2 GHz)  The TCXO can get you do a few ppm or better.
The sensor should let me get about an order of magnitude better, and I
need to measure the temperature anyway, and it's easy to integrate with
the software waveform code.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
De Erroribus Medicorum, R. Bacon, 13th century.
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

Elecraft seems to be having some trouble with the KX3 stability for WSJT on 6m. They also have a temp sensor on the bard, apparently and are also trying this out. I don't know any more about it than that. Don Jim Lux > On 8/30/12 9:33 AM, Bob Camp wrote: >> Hi >> >> The fundamental / third approach is one of several possible ways to >> go. You >> can also run an SC on the B and C modes to get thermometer data. Early >> implementations used a pair of independent blanks, one cut to be a >> good >> thermometer. Some have even gone as far as to mount a thermistor on >> the >> crystal. >> > > I've used the "sensor on oscillator can" technique very recently on a > software defined radio. Over the next year, I'll get some data on how > well it works in use. The idea is to use a low power TCXO and a sensor, > rather than an OCXO, to meet a fairly tight frequency requirement (a few > hundred Hz out of 2 GHz) The TCXO can get you do a few ppm or better. > The sensor should let me get about an order of magnitude better, and I > need to measure the temperature anyway, and it's easy to integrate with > the software waveform code. > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." De Erroribus Medicorum, R. Bacon, 13th century. "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
BC
Bob Camp
Fri, Aug 31, 2012 1:12 AM

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

Bob

On Aug 30, 2012, at 8:01 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/30/12 9:33 AM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to go. You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a good
thermometer. Some have even gone as far as to mount a thermistor on the
crystal.

I've used the "sensor on oscillator can" technique very recently on a software defined radio.  Over the next year, I'll get some data on how well it works in use.  The idea is to use a low power TCXO and a sensor, rather than an OCXO, to meet a fairly tight frequency requirement (a few hundred Hz out of 2 GHz)  The TCXO can get you do a few ppm or better. The sensor should let me get about an order of magnitude better, and I need to measure the temperature anyway, and it's easy to integrate with the software waveform code.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. Bob On Aug 30, 2012, at 8:01 PM, Jim Lux <jimlux@earthlink.net> wrote: > On 8/30/12 9:33 AM, Bob Camp wrote: >> Hi >> >> The fundamental / third approach is one of several possible ways to go. You >> can also run an SC on the B and C modes to get thermometer data. Early >> implementations used a pair of independent blanks, one cut to be a good >> thermometer. Some have even gone as far as to mount a thermistor on the >> crystal. >> > > I've used the "sensor on oscillator can" technique very recently on a software defined radio. Over the next year, I'll get some data on how well it works in use. The idea is to use a low power TCXO and a sensor, rather than an OCXO, to meet a fairly tight frequency requirement (a few hundred Hz out of 2 GHz) The TCXO can get you do a few ppm or better. The sensor should let me get about an order of magnitude better, and I need to measure the temperature anyway, and it's easy to integrate with the software waveform code. > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Fri, Aug 31, 2012 1:29 AM

On 08/31/2012 03:12 AM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

I haven't seen that any of the TCXOs compensates for the temperature
hysteresis. It would be cool if they did.

The ability to handle temperature gradients can be troublesome for both
TCXOs and OCXOs.

TCXOs have become quite good these days.

Cheers,
Magnus

On 08/31/2012 03:12 AM, Bob Camp wrote: > Hi > > If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. I haven't seen that any of the TCXOs compensates for the temperature hysteresis. It would be cool if they did. The ability to handle temperature gradients can be troublesome for both TCXOs and OCXOs. TCXOs have become quite good these days. Cheers, Magnus
BC
Bob Camp
Fri, Aug 31, 2012 1:48 AM

Hi

Some TCXO's have less hysteresis than others…...

Bob

On Aug 30, 2012, at 9:29 PM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

On 08/31/2012 03:12 AM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

I haven't seen that any of the TCXOs compensates for the temperature hysteresis. It would be cool if they did.

The ability to handle temperature gradients can be troublesome for both TCXOs and OCXOs.

TCXOs have become quite good these days.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Some TCXO's have less hysteresis than others…... Bob On Aug 30, 2012, at 9:29 PM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > On 08/31/2012 03:12 AM, Bob Camp wrote: >> Hi >> >> If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. > > I haven't seen that any of the TCXOs compensates for the temperature hysteresis. It would be cool if they did. > > The ability to handle temperature gradients can be troublesome for both TCXOs and OCXOs. > > TCXOs have become quite good these days. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Fri, Aug 31, 2012 2:16 AM

Hi

I think they would do better to just put in a connector to drive it with a TBolt….

Bob

On Aug 30, 2012, at 8:47 PM, Don Latham djl@montana.com wrote:

Elecraft seems to be having some trouble with the KX3 stability for WSJT
on 6m. They also have a temp sensor on the bard, apparently and are also
trying this out. I don't know any more about it than that.
Don

Jim Lux

On 8/30/12 9:33 AM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to
go. You
can also run an SC on the B and C modes to get thermometer data. Early
implementations used a pair of independent blanks, one cut to be a
good
thermometer. Some have even gone as far as to mount a thermistor on
the
crystal.

I've used the "sensor on oscillator can" technique very recently on a
software defined radio.  Over the next year, I'll get some data on how
well it works in use.  The idea is to use a low power TCXO and a sensor,
rather than an OCXO, to meet a fairly tight frequency requirement (a few
hundred Hz out of 2 GHz)  The TCXO can get you do a few ppm or better.
The sensor should let me get about an order of magnitude better, and I
need to measure the temperature anyway, and it's easy to integrate with
the software waveform code.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
De Erroribus Medicorum, R. Bacon, 13th century.
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I think they would do better to just put in a connector to drive it with a TBolt…. Bob On Aug 30, 2012, at 8:47 PM, Don Latham <djl@montana.com> wrote: > Elecraft seems to be having some trouble with the KX3 stability for WSJT > on 6m. They also have a temp sensor on the bard, apparently and are also > trying this out. I don't know any more about it than that. > Don > > Jim Lux >> On 8/30/12 9:33 AM, Bob Camp wrote: >>> Hi >>> >>> The fundamental / third approach is one of several possible ways to >>> go. You >>> can also run an SC on the B and C modes to get thermometer data. Early >>> implementations used a pair of independent blanks, one cut to be a >>> good >>> thermometer. Some have even gone as far as to mount a thermistor on >>> the >>> crystal. >>> >> >> I've used the "sensor on oscillator can" technique very recently on a >> software defined radio. Over the next year, I'll get some data on how >> well it works in use. The idea is to use a low power TCXO and a sensor, >> rather than an OCXO, to meet a fairly tight frequency requirement (a few >> hundred Hz out of 2 GHz) The TCXO can get you do a few ppm or better. >> The sensor should let me get about an order of magnitude better, and I >> need to measure the temperature anyway, and it's easy to integrate with >> the software waveform code. >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > -- > "Neither the voice of authority nor the weight of reason and argument > are as significant as experiment, for thence comes quiet to the mind." > De Erroribus Medicorum, R. Bacon, 13th century. > "If you don't know what it is, don't poke it." > Ghost in the Shell > > > Dr. Don Latham AJ7LL > Six Mile Systems LLP > 17850 Six Mile Road > POB 134 > Huson, MT, 59846 > VOX 406-626-4304 > www.lightningforensics.com > www.sixmilesystems.com > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JL
Jim Lux
Fri, Aug 31, 2012 3:35 AM

On 8/30/12 6:12 PM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

Temp does vary slowly (the radio weighs on the order of 6kg)..

Need to hold spec (in theory) from -20 to +40C.  The oscillator runs
about 10 degrees hotter inside.

About 0.2 ppm from 5C to 40C.  +/- 1ppm worst case over the whole
temperature range.

What I'm also interested in is whether I can compensate a non TCXO 66MHz
CPU clock oscillator (even cheaper, potentially better phase noise with
a higher Q crystal, etc.)

On 8/30/12 6:12 PM, Bob Camp wrote: > Hi > > If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. > Temp does vary slowly (the radio weighs on the order of 6kg).. Need to hold spec (in theory) from -20 to +40C. The oscillator runs about 10 degrees hotter inside. About 0.2 ppm from 5C to 40C. +/- 1ppm worst case over the whole temperature range. What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.)
JL
Jim Lux
Fri, Aug 31, 2012 3:36 AM

On 8/30/12 6:29 PM, Magnus Danielson wrote:

On 08/31/2012 03:12 AM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you
may get your order of magnitude over some range. You might be
surprised at your TCXO. A lot of them are pretty darn good in the
vicinity of room temp. You may already be an order of magnitude past
your ppm or two for fairly normal temperature changes.

I haven't seen that any of the TCXOs compensates for the temperature
hysteresis. It would be cool if they did.

That is the dominant error source for changes over the 5-40C.. about 0.2
ppm difference when going up from -55 to +85 and going down over the
same range.

The ability to handle temperature gradients can be troublesome for both
TCXOs and OCXOs.

TCXOs have become quite good these days.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

On 8/30/12 6:29 PM, Magnus Danielson wrote: > On 08/31/2012 03:12 AM, Bob Camp wrote: >> Hi >> >> If the temperature is varying slowly *and* there are no gradients you >> may get your order of magnitude over some range. You might be >> surprised at your TCXO. A lot of them are pretty darn good in the >> vicinity of room temp. You may already be an order of magnitude past >> your ppm or two for fairly normal temperature changes. > > I haven't seen that any of the TCXOs compensates for the temperature > hysteresis. It would be cool if they did. That is the dominant error source for changes over the 5-40C.. about 0.2 ppm difference when going up from -55 to +85 and going down over the same range. > > The ability to handle temperature gradients can be troublesome for both > TCXOs and OCXOs. > > TCXOs have become quite good these days. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Fri, Aug 31, 2012 4:04 AM

Hi

A standard clock crystal is pretty much junk. as far as temperature performance is concerned. Even a cheap TCXO is likely to be pretty good over 25 C +/- 10C.

Bob

On Aug 30, 2012, at 11:35 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/30/12 6:12 PM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

Temp does vary slowly (the radio weighs on the order of 6kg)..

Need to hold spec (in theory) from -20 to +40C.  The oscillator runs about 10 degrees hotter inside.

About 0.2 ppm from 5C to 40C.  +/- 1ppm worst case over the whole temperature range.

What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi A standard clock crystal is pretty much junk. as far as temperature performance is concerned. Even a cheap TCXO is likely to be pretty good over 25 C +/- 10C. Bob On Aug 30, 2012, at 11:35 PM, Jim Lux <jimlux@earthlink.net> wrote: > On 8/30/12 6:12 PM, Bob Camp wrote: >> Hi >> >> If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. >> > > Temp does vary slowly (the radio weighs on the order of 6kg).. > > Need to hold spec (in theory) from -20 to +40C. The oscillator runs about 10 degrees hotter inside. > > About 0.2 ppm from 5C to 40C. +/- 1ppm worst case over the whole temperature range. > > What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.) > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JL
Jim Lux
Fri, Aug 31, 2012 4:10 AM

On 8/30/12 9:04 PM, Bob Camp wrote:

Hi

A standard clock crystal is pretty much junk. as far as temperature performance is concerned. Even a cheap TCXO is likely to be pretty good over 25 C +/- 10C.

yes.. but, say, one had a decent SC cut with good phase noise
properties, but large (but repeatable) temperature characteristics.

Can I get good accuracy AND good phase noise, for less hassle/power/size
than an OCXO?

My radio has a TCXO and a clock oscillator, so the clock oscillator is
my test article for the temperature compensation scheme..

Bob

On Aug 30, 2012, at 11:35 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/30/12 6:12 PM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

Temp does vary slowly (the radio weighs on the order of 6kg)..

Need to hold spec (in theory) from -20 to +40C.  The oscillator runs about 10 degrees hotter inside.

About 0.2 ppm from 5C to 40C.  +/- 1ppm worst case over the whole temperature range.

What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

On 8/30/12 9:04 PM, Bob Camp wrote: > Hi > > A standard clock crystal is pretty much junk. as far as temperature performance is concerned. Even a cheap TCXO is likely to be pretty good over 25 C +/- 10C. > yes.. but, say, one had a decent SC cut with good phase noise properties, but large (but repeatable) temperature characteristics. Can I get good accuracy AND good phase noise, for less hassle/power/size than an OCXO? My radio has a TCXO and a clock oscillator, so the clock oscillator is my test article for the temperature compensation scheme.. > Bob > > On Aug 30, 2012, at 11:35 PM, Jim Lux <jimlux@earthlink.net> wrote: > >> On 8/30/12 6:12 PM, Bob Camp wrote: >>> Hi >>> >>> If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. >>> >> >> Temp does vary slowly (the radio weighs on the order of 6kg).. >> >> Need to hold spec (in theory) from -20 to +40C. The oscillator runs about 10 degrees hotter inside. >> >> About 0.2 ppm from 5C to 40C. +/- 1ppm worst case over the whole temperature range. >> >> What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.) >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
DL
Don Latham
Fri, Aug 31, 2012 5:34 AM

Bob: Great minds run in the same track :-)
Don

Bob Camp

Hi

I think they would do better to just put in a connector to drive it with
a TBolt….

Bob

On Aug 30, 2012, at 8:47 PM, Don Latham djl@montana.com wrote:

Elecraft seems to be having some trouble with the KX3 stability for
WSJT
on 6m. They also have a temp sensor on the bard, apparently and are
also
trying this out. I don't know any more about it than that.
Don

Jim Lux

On 8/30/12 9:33 AM, Bob Camp wrote:

Hi

The fundamental / third approach is one of several possible ways to
go. You
can also run an SC on the B and C modes to get thermometer data.
Early
implementations used a pair of independent blanks, one cut to be a
good
thermometer. Some have even gone as far as to mount a thermistor on
the
crystal.

I've used the "sensor on oscillator can" technique very recently on a
software defined radio.  Over the next year, I'll get some data on
how
well it works in use.  The idea is to use a low power TCXO and a
sensor,
rather than an OCXO, to meet a fairly tight frequency requirement (a
few
hundred Hz out of 2 GHz)  The TCXO can get you do a few ppm or
better.
The sensor should let me get about an order of magnitude better, and
I
need to measure the temperature anyway, and it's easy to integrate
with
the software waveform code.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
De Erroribus Medicorum, R. Bacon, 13th century.
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
"Neither the voice of authority nor the weight of reason and argument
are as significant as experiment, for thence comes quiet to the mind."
De Erroribus Medicorum, R. Bacon, 13th century.
"If you don't know what it is, don't poke it."
Ghost in the Shell

Dr. Don Latham AJ7LL
Six Mile Systems LLP
17850 Six Mile Road
POB 134
Huson, MT, 59846
VOX 406-626-4304
www.lightningforensics.com
www.sixmilesystems.com

Bob: Great minds run in the same track :-) Don Bob Camp > Hi > > I think they would do better to just put in a connector to drive it with > a TBolt…. > > Bob > > On Aug 30, 2012, at 8:47 PM, Don Latham <djl@montana.com> wrote: > >> Elecraft seems to be having some trouble with the KX3 stability for >> WSJT >> on 6m. They also have a temp sensor on the bard, apparently and are >> also >> trying this out. I don't know any more about it than that. >> Don >> >> Jim Lux >>> On 8/30/12 9:33 AM, Bob Camp wrote: >>>> Hi >>>> >>>> The fundamental / third approach is one of several possible ways to >>>> go. You >>>> can also run an SC on the B and C modes to get thermometer data. >>>> Early >>>> implementations used a pair of independent blanks, one cut to be a >>>> good >>>> thermometer. Some have even gone as far as to mount a thermistor on >>>> the >>>> crystal. >>>> >>> >>> I've used the "sensor on oscillator can" technique very recently on a >>> software defined radio. Over the next year, I'll get some data on >>> how >>> well it works in use. The idea is to use a low power TCXO and a >>> sensor, >>> rather than an OCXO, to meet a fairly tight frequency requirement (a >>> few >>> hundred Hz out of 2 GHz) The TCXO can get you do a few ppm or >>> better. >>> The sensor should let me get about an order of magnitude better, and >>> I >>> need to measure the temperature anyway, and it's easy to integrate >>> with >>> the software waveform code. >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> >> -- >> "Neither the voice of authority nor the weight of reason and argument >> are as significant as experiment, for thence comes quiet to the mind." >> De Erroribus Medicorum, R. Bacon, 13th century. >> "If you don't know what it is, don't poke it." >> Ghost in the Shell >> >> >> Dr. Don Latham AJ7LL >> Six Mile Systems LLP >> 17850 Six Mile Road >> POB 134 >> Huson, MT, 59846 >> VOX 406-626-4304 >> www.lightningforensics.com >> www.sixmilesystems.com >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > -- "Neither the voice of authority nor the weight of reason and argument are as significant as experiment, for thence comes quiet to the mind." De Erroribus Medicorum, R. Bacon, 13th century. "If you don't know what it is, don't poke it." Ghost in the Shell Dr. Don Latham AJ7LL Six Mile Systems LLP 17850 Six Mile Road POB 134 Huson, MT, 59846 VOX 406-626-4304 www.lightningforensics.com www.sixmilesystems.com
BC
Bob Camp
Fri, Aug 31, 2012 2:06 PM

Hi

An SC is going to have it's temperature curve centered up around 95C or so. If it's been cut as an OCXO crystal the turns will be up there as well. By the time it gets to room temp, the delta F / delta T is moving mighty fast. Think in terms of multiple ppm/C. A typical cell phone TCXO crystal is in the sub 0.1 ppm/C range in the vicinity of room temp.
In addition, SC's are pretty hard to pull. For a normal TCXO (no DDS) something > 40 ppm of range would be needed. By the time you get the wide tune stuff in the circuit, the phase noise isn't going to be anything special.

Bob

On Aug 31, 2012, at 12:10 AM, Jim Lux jimlux@earthlink.net wrote:

On 8/30/12 9:04 PM, Bob Camp wrote:

Hi

A standard clock crystal is pretty much junk. as far as temperature performance is concerned. Even a cheap TCXO is likely to be pretty good over 25 C +/- 10C.

yes.. but, say, one had a decent SC cut with good phase noise properties, but large (but repeatable) temperature characteristics.

Can I get good accuracy AND good phase noise, for less hassle/power/size than an OCXO?

My radio has a TCXO and a clock oscillator, so the clock oscillator is my test article for the temperature compensation scheme..

Bob

On Aug 30, 2012, at 11:35 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/30/12 6:12 PM, Bob Camp wrote:

Hi

If the temperature is varying slowly and there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes.

Temp does vary slowly (the radio weighs on the order of 6kg)..

Need to hold spec (in theory) from -20 to +40C.  The oscillator runs about 10 degrees hotter inside.

About 0.2 ppm from 5C to 40C.  +/- 1ppm worst case over the whole temperature range.

What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi An SC is going to have it's temperature curve centered up around 95C or so. If it's been cut as an OCXO crystal the turns will be up there as well. By the time it gets to room temp, the delta F / delta T is moving mighty fast. Think in terms of multiple ppm/C. A typical cell phone TCXO crystal is in the sub 0.1 ppm/C range in the vicinity of room temp. In addition, SC's are pretty hard to pull. For a normal TCXO (no DDS) something > 40 ppm of range would be needed. By the time you get the wide tune stuff in the circuit, the phase noise isn't going to be anything special. Bob On Aug 31, 2012, at 12:10 AM, Jim Lux <jimlux@earthlink.net> wrote: > On 8/30/12 9:04 PM, Bob Camp wrote: >> Hi >> >> A standard clock crystal is pretty much junk. as far as temperature performance is concerned. Even a cheap TCXO is likely to be pretty good over 25 C +/- 10C. >> > > yes.. but, say, one had a decent SC cut with good phase noise properties, but large (but repeatable) temperature characteristics. > > Can I get good accuracy AND good phase noise, for less hassle/power/size than an OCXO? > > My radio has a TCXO and a clock oscillator, so the clock oscillator is my test article for the temperature compensation scheme.. > > > >> Bob >> >> On Aug 30, 2012, at 11:35 PM, Jim Lux <jimlux@earthlink.net> wrote: >> >>> On 8/30/12 6:12 PM, Bob Camp wrote: >>>> Hi >>>> >>>> If the temperature is varying slowly *and* there are no gradients you may get your order of magnitude over some range. You might be surprised at your TCXO. A lot of them are pretty darn good in the vicinity of room temp. You may already be an order of magnitude past your ppm or two for fairly normal temperature changes. >>>> >>> >>> Temp does vary slowly (the radio weighs on the order of 6kg).. >>> >>> Need to hold spec (in theory) from -20 to +40C. The oscillator runs about 10 degrees hotter inside. >>> >>> About 0.2 ppm from 5C to 40C. +/- 1ppm worst case over the whole temperature range. >>> >>> What I'm also interested in is whether I can compensate a non TCXO 66MHz CPU clock oscillator (even cheaper, potentially better phase noise with a higher Q crystal, etc.) >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JL
Jim Lux
Sat, Sep 1, 2012 12:55 AM

On 8/31/12 7:06 AM, Bob Camp wrote:

Hi

An SC is going to have it's temperature curve centered up around 95C
or so. If it's been cut as an OCXO crystal the turns will be up there
as well. By the time it gets to room temp, the delta F / delta T is
moving mighty fast. Think in terms of multiple ppm/C. A typical cell
phone TCXO crystal is in the sub 0.1 ppm/C range in the vicinity of
room temp. In addition, SC's are pretty hard to pull. For a normal
TCXO (no DDS) something > 40 ppm of range would be needed. By the
time you get the wide tune stuff in the circuit, the phase noise
isn't going to be anything special.

Bob

We're not pulling the crystal.. we just let it sit.
What we do is change an NCO in the software radio implementation.
Typically, you have a coarse tune with a PLL in steps of 500kHz-1MHz,
and then you do the fine tune in software with a digital
mixer/downconverter and an NCO (which also is part of the carrier
tracking loop).

On 8/31/12 7:06 AM, Bob Camp wrote: > Hi > > An SC is going to have it's temperature curve centered up around 95C > or so. If it's been cut as an OCXO crystal the turns will be up there > as well. By the time it gets to room temp, the delta F / delta T is > moving mighty fast. Think in terms of multiple ppm/C. A typical cell > phone TCXO crystal is in the sub 0.1 ppm/C range in the vicinity of > room temp. In addition, SC's are pretty hard to pull. For a normal > TCXO (no DDS) something > 40 ppm of range would be needed. By the > time you get the wide tune stuff in the circuit, the phase noise > isn't going to be anything special. > > Bob > We're not pulling the crystal.. we just let it sit. What we do is change an NCO in the software radio implementation. Typically, you have a coarse tune with a PLL in steps of 500kHz-1MHz, and then you do the fine tune in software with a digital mixer/downconverter and an NCO (which also is part of the carrier tracking loop).
BC
Bob Camp
Sat, Sep 1, 2012 2:16 AM

Hi

Ok, that gets you back to the basics of really major delta F / delta T slopes.

Bob

On Aug 31, 2012, at 8:55 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/31/12 7:06 AM, Bob Camp wrote:

Hi

An SC is going to have it's temperature curve centered up around 95C
or so. If it's been cut as an OCXO crystal the turns will be up there
as well. By the time it gets to room temp, the delta F / delta T is
moving mighty fast. Think in terms of multiple ppm/C. A typical cell
phone TCXO crystal is in the sub 0.1 ppm/C range in the vicinity of
room temp. In addition, SC's are pretty hard to pull. For a normal
TCXO (no DDS) something > 40 ppm of range would be needed. By the
time you get the wide tune stuff in the circuit, the phase noise
isn't going to be anything special.

Bob

We're not pulling the crystal.. we just let it sit.
What we do is change an NCO in the software radio implementation. Typically, you have a coarse tune with a PLL in steps of 500kHz-1MHz, and then you do the fine tune in software with a digital mixer/downconverter and an NCO (which also is part of the carrier tracking loop).


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Ok, that gets you back to the basics of really major delta F / delta T slopes. Bob On Aug 31, 2012, at 8:55 PM, Jim Lux <jimlux@earthlink.net> wrote: > On 8/31/12 7:06 AM, Bob Camp wrote: >> Hi >> >> An SC is going to have it's temperature curve centered up around 95C >> or so. If it's been cut as an OCXO crystal the turns will be up there >> as well. By the time it gets to room temp, the delta F / delta T is >> moving mighty fast. Think in terms of multiple ppm/C. A typical cell >> phone TCXO crystal is in the sub 0.1 ppm/C range in the vicinity of >> room temp. In addition, SC's are pretty hard to pull. For a normal >> TCXO (no DDS) something > 40 ppm of range would be needed. By the >> time you get the wide tune stuff in the circuit, the phase noise >> isn't going to be anything special. >> >> Bob >> > > We're not pulling the crystal.. we just let it sit. > What we do is change an NCO in the software radio implementation. Typically, you have a coarse tune with a PLL in steps of 500kHz-1MHz, and then you do the fine tune in software with a digital mixer/downconverter and an NCO (which also is part of the carrier tracking loop). > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JL
Jim Lux
Sat, Sep 1, 2012 3:26 AM

On 8/31/12 7:16 PM, Bob Camp wrote:

Hi

Ok, that gets you back to the basics of really major delta F / delta T slopes.

Bob

yeah.. but as long as you know what the curve is.. the NCO has a huge
range (after all, we already have to tune over >500 kHz...a few hundred
Hz isn't a big deal.

A more complex problem is doing insitu calibration from, e.g., GPS
signals or from some externally received frequency reference. (since the
radio in question can also work as a GPS receiver, eventually).

On 8/31/12 7:16 PM, Bob Camp wrote: > Hi > > Ok, that gets you back to the basics of really major delta F / delta T slopes. > > Bob > yeah.. but as long as you know what the curve is.. the NCO has a huge range (after all, we already have to tune over >500 kHz...a few hundred Hz isn't a big deal. A more complex problem is doing insitu calibration from, e.g., GPS signals or from some externally received frequency reference. (since the radio in question can also work as a GPS receiver, eventually).
BC
Bob Camp
Sat, Sep 1, 2012 3:32 PM

Hi

Observing a curve and being able to compensate it are often two different things. Hysteresis is one very obvious example. Another is simple sensor lag. A some what less obvious one is that the temperature performance is also influenced by the rate of change in temperature.

Here's another thing to consider:

If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 stability at one second - You will need to either have a rate of change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to compensate for some pretty small changes. That of course makes a bunch of assumptions ….

Bob

On Aug 31, 2012, at 11:26 PM, Jim Lux jimlux@earthlink.net wrote:

On 8/31/12 7:16 PM, Bob Camp wrote:

Hi

Ok, that gets you back to the basics of really major delta F / delta T slopes.

Bob

yeah.. but as long as you know what the curve is.. the NCO has a huge range (after all, we already have to tune over >500 kHz...a few hundred Hz isn't a big deal.

A more complex problem is doing insitu calibration from, e.g., GPS signals or from some externally received frequency reference. (since the radio in question can also work as a GPS receiver, eventually).


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Observing a curve and being able to compensate it are often two different things. Hysteresis is one very obvious example. Another is simple sensor lag. A some what less obvious one is that the temperature performance is also influenced by the rate of change in temperature. Here's another thing to consider: If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 stability at one second - You will need to either have a rate of change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to compensate for some pretty small changes. That of course makes a bunch of assumptions …. Bob On Aug 31, 2012, at 11:26 PM, Jim Lux <jimlux@earthlink.net> wrote: > On 8/31/12 7:16 PM, Bob Camp wrote: >> Hi >> >> Ok, that gets you back to the basics of really major delta F / delta T slopes. >> >> Bob >> > yeah.. but as long as you know what the curve is.. the NCO has a huge range (after all, we already have to tune over >500 kHz...a few hundred Hz isn't a big deal. > > > A more complex problem is doing insitu calibration from, e.g., GPS signals or from some externally received frequency reference. (since the radio in question can also work as a GPS receiver, eventually). > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JL
Jim Lux
Sat, Sep 1, 2012 3:53 PM

On 9/1/12 8:32 AM, Bob Camp wrote:

Hi

Observing a curve and being able to compensate it are often two different things. Hysteresis is one very obvious example. Another is simple sensor lag. A some what less obvious one is that the temperature performance is also influenced by the rate of change in temperature.

Here's another thing to consider:

If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 stability at one second - You will need to either have a rate of change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to compensate for some pretty small changes. That of course makes a bunch of assumptions ….

In this application,  the requirement for frequency accuracy has to do
with initial acquisition.. that is, you want the signal (or receiver
tuning) to be within some few hundred Hz of where it's expected to be
(because the receiver is narrow band).

The ground station typically has a Doppler predict based on orbit
knowledge, that predict has some uncertainty.  Added to the radio
frequency uncertainty.  (SNUG - Space Network User Guide, has more info)

Once you've acquired, the receiver and ground station will track (i.e.
the ground station puts in the estimated Doppler, so all you're really
tracking is the variation in the local oscillator).  (for a LEO
satellite at 2.3 GHz, the 7km/s orbital velocity already puts tens of
kHz variation on it)

(and this completely neglects that a modern radio could use something
like an FFT for acquisition)

Temperature changes are pretty slow.. I'm seeing 5-10 degree cyclical
variation over 90-100 minutes.  Actually, the bigger change is during
the warm up transient, going from off and cold to on and warm over 10
minutes or so.

In other applications, where you're not going in and out of the sun
every revolution (i.e. deep space, rather than LEO) and you were
interested in Allan deviation type measurements for gravity science
(where we're looking for 1E-13 over 100 sec sort of performance), what
we'd probably do is warm up early.. Turn it on, compensate based on the
measured temperature, and then hold the compensation fixed during the
measurement, letting the ground worry about the apparent frequency
change due to Doppler.  We'd have a high quality narrow band signal,
just at an unknown (but reasonably stable) frequency.  What the science
team is usually interested in is small relative changes in phase &
amplitude(occultations) or in small changes in frequency (Doppler, for
gravity science).

(we regularly measure velocity to cm/sec precision for outer planet
orbiters like Cassini, Juno, etc.)

On 9/1/12 8:32 AM, Bob Camp wrote: > Hi > > Observing a curve and being able to compensate it are often two different things. Hysteresis is one very obvious example. Another is simple sensor lag. A some what less obvious one is that the temperature performance is also influenced by the rate of change in temperature. > > Here's another thing to consider: > > If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 stability at one second - You will need to either have a rate of change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to compensate for some pretty small changes. That of course makes a bunch of assumptions …. > > In this application, the requirement for frequency accuracy has to do with initial acquisition.. that is, you want the signal (or receiver tuning) to be within some few hundred Hz of where it's expected to be (because the receiver is narrow band). The ground station typically has a Doppler predict based on orbit knowledge, that predict has some uncertainty. Added to the radio frequency uncertainty. (SNUG - Space Network User Guide, has more info) Once you've acquired, the receiver and ground station will track (i.e. the ground station puts in the estimated Doppler, so all you're really tracking is the variation in the local oscillator). (for a LEO satellite at 2.3 GHz, the 7km/s orbital velocity already puts tens of kHz variation on it) (and this completely neglects that a modern radio could use something like an FFT for acquisition) Temperature changes are pretty slow.. I'm seeing 5-10 degree cyclical variation over 90-100 minutes. Actually, the bigger change is during the warm up transient, going from off and cold to on and warm over 10 minutes or so. In other applications, where you're not going in and out of the sun every revolution (i.e. deep space, rather than LEO) and you were interested in Allan deviation type measurements for gravity science (where we're looking for 1E-13 over 100 sec sort of performance), what we'd probably do is warm up early.. Turn it on, compensate based on the measured temperature, and then hold the compensation fixed during the measurement, letting the ground worry about the apparent frequency change due to Doppler. We'd have a high quality narrow band signal, just at an unknown (but reasonably stable) frequency. What the science team is usually interested in is small relative changes in phase & amplitude(occultations) or in small changes in frequency (Doppler, for gravity science). (we regularly measure velocity to cm/sec precision for outer planet orbiters like Cassini, Juno, etc.)
BC
Bob Camp
Sat, Sep 1, 2012 6:00 PM

Hi

I suspect that a good IT cut would probably do better than an SC in either application. In the deep space situation, a copper slug in a dewar sounds like a reasonable addition to the design.

Bob

On Sep 1, 2012, at 11:53 AM, Jim Lux jimlux@earthlink.net wrote:

On 9/1/12 8:32 AM, Bob Camp wrote:

Hi

Observing a curve and being able to compensate it are often two different things. Hysteresis is one very obvious example. Another is simple sensor lag. A some what less obvious one is that the temperature performance is also influenced by the rate of change in temperature.

Here's another thing to consider:

If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 stability at one second - You will need to either have a rate of change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to compensate for some pretty small changes. That of course makes a bunch of assumptions ….

In this application,  the requirement for frequency accuracy has to do with initial acquisition.. that is, you want the signal (or receiver tuning) to be within some few hundred Hz of where it's expected to be (because the receiver is narrow band).

The ground station typically has a Doppler predict based on orbit knowledge, that predict has some uncertainty.  Added to the radio frequency uncertainty.  (SNUG - Space Network User Guide, has more info)

Once you've acquired, the receiver and ground station will track (i.e. the ground station puts in the estimated Doppler, so all you're really tracking is the variation in the local oscillator).  (for a LEO satellite at 2.3 GHz, the 7km/s orbital velocity already puts tens of kHz variation on it)

(and this completely neglects that a modern radio could use something like an FFT for acquisition)

Temperature changes are pretty slow.. I'm seeing 5-10 degree cyclical variation over 90-100 minutes.  Actually, the bigger change is during the warm up transient, going from off and cold to on and warm over 10 minutes or so.

In other applications, where you're not going in and out of the sun every revolution (i.e. deep space, rather than LEO) and you were interested in Allan deviation type measurements for gravity science (where we're looking for 1E-13 over 100 sec sort of performance), what we'd probably do is warm up early.. Turn it on, compensate based on the measured temperature, and then hold the compensation fixed during the measurement, letting the ground worry about the apparent frequency change due to Doppler.  We'd have a high quality narrow band signal, just at an unknown (but reasonably stable) frequency.  What the science team is usually interested in is small relative changes in phase & amplitude(occultations) or in small changes in frequency (Doppler, for gravity science).

(we regularly measure velocity to cm/sec precision for outer planet orbiters like Cassini, Juno, etc.)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I suspect that a good IT cut would probably do better than an SC in either application. In the deep space situation, a copper slug in a dewar sounds like a reasonable addition to the design. Bob On Sep 1, 2012, at 11:53 AM, Jim Lux <jimlux@earthlink.net> wrote: > On 9/1/12 8:32 AM, Bob Camp wrote: >> Hi >> >> Observing a curve and being able to compensate it are often two different things. Hysteresis is one very obvious example. Another is simple sensor lag. A some what less obvious one is that the temperature performance is also influenced by the rate of change in temperature. >> >> Here's another thing to consider: >> >> If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 stability at one second - You will need to either have a rate of change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to compensate for some pretty small changes. That of course makes a bunch of assumptions …. >> >> > > In this application, the requirement for frequency accuracy has to do with initial acquisition.. that is, you want the signal (or receiver tuning) to be within some few hundred Hz of where it's expected to be (because the receiver is narrow band). > > > The ground station typically has a Doppler predict based on orbit knowledge, that predict has some uncertainty. Added to the radio frequency uncertainty. (SNUG - Space Network User Guide, has more info) > > Once you've acquired, the receiver and ground station will track (i.e. the ground station puts in the estimated Doppler, so all you're really tracking is the variation in the local oscillator). (for a LEO satellite at 2.3 GHz, the 7km/s orbital velocity already puts tens of kHz variation on it) > > (and this completely neglects that a modern radio could use something like an FFT for acquisition) > > Temperature changes are pretty slow.. I'm seeing 5-10 degree cyclical variation over 90-100 minutes. Actually, the bigger change is during the warm up transient, going from off and cold to on and warm over 10 minutes or so. > > In other applications, where you're not going in and out of the sun every revolution (i.e. deep space, rather than LEO) and you were interested in Allan deviation type measurements for gravity science (where we're looking for 1E-13 over 100 sec sort of performance), what we'd probably do is warm up early.. Turn it on, compensate based on the measured temperature, and then hold the compensation fixed during the measurement, letting the ground worry about the apparent frequency change due to Doppler. We'd have a high quality narrow band signal, just at an unknown (but reasonably stable) frequency. What the science team is usually interested in is small relative changes in phase & amplitude(occultations) or in small changes in frequency (Doppler, for gravity science). > > (we regularly measure velocity to cm/sec precision for outer planet orbiters like Cassini, Juno, etc.) > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Sat, Sep 1, 2012 9:01 PM

On 09/01/2012 05:53 PM, Jim Lux wrote:

On 9/1/12 8:32 AM, Bob Camp wrote:

Hi

Observing a curve and being able to compensate it are often two
different things. Hysteresis is one very obvious example. Another is
simple sensor lag. A some what less obvious one is that the
temperature performance is also influenced by the rate of change in
temperature.

Here's another thing to consider:

If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11
stability at one second - You will need to either have a rate of
change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to
compensate for some pretty small changes. That of course makes a bunch
of assumptions ….

In this application, the requirement for frequency accuracy has to do
with initial acquisition.. that is, you want the signal (or receiver
tuning) to be within some few hundred Hz of where it's expected to be
(because the receiver is narrow band).

The ground station typically has a Doppler predict based on orbit
knowledge, that predict has some uncertainty. Added to the radio
frequency uncertainty. (SNUG - Space Network User Guide, has more info)

Once you've acquired, the receiver and ground station will track (i.e.
the ground station puts in the estimated Doppler, so all you're really
tracking is the variation in the local oscillator). (for a LEO satellite
at 2.3 GHz, the 7km/s orbital velocity already puts tens of kHz
variation on it)

(and this completely neglects that a modern radio could use something
like an FFT for acquisition)

Temperature changes are pretty slow.. I'm seeing 5-10 degree cyclical
variation over 90-100 minutes. Actually, the bigger change is during the
warm up transient, going from off and cold to on and warm over 10
minutes or so.

In other applications, where you're not going in and out of the sun
every revolution (i.e. deep space, rather than LEO) and you were
interested in Allan deviation type measurements for gravity science
(where we're looking for 1E-13 over 100 sec sort of performance), what
we'd probably do is warm up early.. Turn it on, compensate based on the
measured temperature, and then hold the compensation fixed during the
measurement, letting the ground worry about the apparent frequency
change due to Doppler. We'd have a high quality narrow band signal, just
at an unknown (but reasonably stable) frequency. What the science team
is usually interested in is small relative changes in phase &
amplitude(occultations) or in small changes in frequency (Doppler, for
gravity science).

(we regularly measure velocity to cm/sec precision for outer planet
orbiters like Cassini, Juno, etc.)

If you can make reasonable predictions of the heating and cooling
profile, you could use that and hopefully gain a decade or so, and any
slew in detectors and would mostly affect the remaining error.

You could also make use of uplink carrier and GPS to improve the model
state of your crystal. That way the predicted and feed forward values
could be kept fairly well adapted. Naturally, you would want a fall-back
scenario to back out for wider offset search if you failed to maintain
the model. You don't want to loose your bird due to temporary system
failure. Another alternative would be to allow for the downlink
frequency to be a hint for the uplink where the frequency is such that
you can "pull in" the bird if need to.

Cheers,
Magnus

On 09/01/2012 05:53 PM, Jim Lux wrote: > On 9/1/12 8:32 AM, Bob Camp wrote: >> Hi >> >> Observing a curve and being able to compensate it are often two >> different things. Hysteresis is one very obvious example. Another is >> simple sensor lag. A some what less obvious one is that the >> temperature performance is also influenced by the rate of change in >> temperature. >> >> Here's another thing to consider: >> >> If your crystal is running 3 ppm / C, and you are after 3.0 x 10^-11 >> stability at one second - You will need to either have a rate of >> change at ~ 1x10^-5 C/sec (0.6 mC / min) or you will need to >> compensate for some pretty small changes. That of course makes a bunch >> of assumptions …. >> >> > > In this application, the requirement for frequency accuracy has to do > with initial acquisition.. that is, you want the signal (or receiver > tuning) to be within some few hundred Hz of where it's expected to be > (because the receiver is narrow band). > > > The ground station typically has a Doppler predict based on orbit > knowledge, that predict has some uncertainty. Added to the radio > frequency uncertainty. (SNUG - Space Network User Guide, has more info) > > Once you've acquired, the receiver and ground station will track (i.e. > the ground station puts in the estimated Doppler, so all you're really > tracking is the variation in the local oscillator). (for a LEO satellite > at 2.3 GHz, the 7km/s orbital velocity already puts tens of kHz > variation on it) > > (and this completely neglects that a modern radio could use something > like an FFT for acquisition) > > Temperature changes are pretty slow.. I'm seeing 5-10 degree cyclical > variation over 90-100 minutes. Actually, the bigger change is during the > warm up transient, going from off and cold to on and warm over 10 > minutes or so. > > In other applications, where you're not going in and out of the sun > every revolution (i.e. deep space, rather than LEO) and you were > interested in Allan deviation type measurements for gravity science > (where we're looking for 1E-13 over 100 sec sort of performance), what > we'd probably do is warm up early.. Turn it on, compensate based on the > measured temperature, and then hold the compensation fixed during the > measurement, letting the ground worry about the apparent frequency > change due to Doppler. We'd have a high quality narrow band signal, just > at an unknown (but reasonably stable) frequency. What the science team > is usually interested in is small relative changes in phase & > amplitude(occultations) or in small changes in frequency (Doppler, for > gravity science). > > (we regularly measure velocity to cm/sec precision for outer planet > orbiters like Cassini, Juno, etc.) If you can make reasonable predictions of the heating and cooling profile, you could use that and hopefully gain a decade or so, and any slew in detectors and would mostly affect the remaining error. You could also make use of uplink carrier and GPS to improve the model state of your crystal. That way the predicted and feed forward values could be kept fairly well adapted. Naturally, you would want a fall-back scenario to back out for wider offset search if you failed to maintain the model. You don't want to loose your bird due to temporary system failure. Another alternative would be to allow for the downlink frequency to be a hint for the uplink where the frequency is such that you can "pull in" the bird if need to. Cheers, Magnus
JL
Jim Lux
Sun, Sep 2, 2012 2:41 AM

On 9/1/12 11:00 AM, Bob Camp wrote:

Hi

I suspect that a good IT cut would probably do better than an SC in either application. In the deep space situation, a copper slug in a dewar sounds like a reasonable addition to the design.

If you're going to do dewars, then you're talking USOs for which the
technology is quite well developed with ovens, etc.

But that's a kilo, several watts, and a liter.

What you want is something that is better than, say, 0.1 ppm, that is
comparable in mass/power/volume/cost to an existing 1-5 ppm TCXO (about
1x2x0.5 cm)

On 9/1/12 11:00 AM, Bob Camp wrote: > Hi > > I suspect that a good IT cut would probably do better than an SC in either application. In the deep space situation, a copper slug in a dewar sounds like a reasonable addition to the design. > If you're going to do dewars, then you're talking USOs for which the technology is quite well developed with ovens, etc. But that's a kilo, several watts, and a liter. What you want is something that is better than, say, 0.1 ppm, that is comparable in mass/power/volume/cost to an existing 1-5 ppm TCXO (about 1x2x0.5 cm)
BC
Bob Camp
Sun, Sep 2, 2012 2:52 AM

Hi

True on the volume and weight. Not as much power as an OCXO since it's passive. At 1x 2 x 0.5 you could fix the power and weight by using an EMXO. You still would have more power than the TCXO, but no were near as much as the couple watts a USO uses.

Bob

Bob

On Sep 1, 2012, at 10:41 PM, Jim Lux jimlux@earthlink.net wrote:

On 9/1/12 11:00 AM, Bob Camp wrote:

Hi

I suspect that a good IT cut would probably do better than an SC in either application. In the deep space situation, a copper slug in a dewar sounds like a reasonable addition to the design.

If you're going to do dewars, then you're talking USOs for which the technology is quite well developed with ovens, etc.

But that's a kilo, several watts, and a liter.

What you want is something that is better than, say, 0.1 ppm, that is comparable in mass/power/volume/cost to an existing 1-5 ppm TCXO (about 1x2x0.5 cm)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi True on the volume and weight. Not as much power as an OCXO since it's passive. At 1x 2 x 0.5 you could fix the power and weight by using an EMXO. You still would have more power than the TCXO, but no were near as much as the couple watts a USO uses. Bob Bob On Sep 1, 2012, at 10:41 PM, Jim Lux <jimlux@earthlink.net> wrote: > On 9/1/12 11:00 AM, Bob Camp wrote: >> Hi >> >> I suspect that a good IT cut would probably do better than an SC in either application. In the deep space situation, a copper slug in a dewar sounds like a reasonable addition to the design. >> > > If you're going to do dewars, then you're talking USOs for which the technology is quite well developed with ovens, etc. > > But that's a kilo, several watts, and a liter. > > What you want is something that is better than, say, 0.1 ppm, that is comparable in mass/power/volume/cost to an existing 1-5 ppm TCXO (about 1x2x0.5 cm) > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.