Discussion of precise voltage measurement

View all threads

Re: [volt-nuts] HP 3458A

FS
Fred Schneider
Sun, Aug 7, 2011 4:50 AM

Calibration, is only testing and noting the deviation from the standard used. To adjust the meter to this standard is an extra service. And indeed shipping, other air humidity, other temperature ect and time will change this. I have a guildline cabinet. The papers it was calibrated every year in the 79's and 80's. I have the reports ( and some copy of the bills) because it came with all documentation . I got it as a gift from an internal calibration lab of a very big company. I was lookimg on the net to search info and I wrote some firms like guildline to find out more about these wonderfull cabinets. Then I heard that calibrating such a cabinet ( that was, if I lived in the USA, but over here it must be allso posdible I guess) was still possible and  transportation alone was 750 dollar)

Fred PA4TIM

Op 7 aug. 2011 om 02:13 heeft WB6BNQ wb6bnq@cox.net het volgende geschreven:

This response illustrates the absurdity in the whole process.  The so-called
"Golden Calibration" is only of value in shifting from a fixed standard to an
instrument that can measure "in-between" values to a high degree if done within
minutes of the set up.  It is called the transfer ratio and is the 24 hour
specification.

The absurdity to think your going spend additional money for the "Golden" part
and get something is total BS.  Even the STE/9000 is the value at the time of
calibration.  Bouncing around in shipping could certainly invalidate that value.

Such antics could best be described as the cat chasing his tail.

Bill....WB6BNQ

gbusg wrote:

Hi Brooke,

MU at 10Vdc for the "Golden Calibration" is in the range +/- 0.2 ppm, IIRC.

MU at 10Vdc for the "STE/9000 Calibration" is in the range +/- 3.6 ppm,
IIRC.

The "Golden Calibration" verifies the 3458A to its 24 hour specs.

The "STE/9000 Calibration" verifies the 3458A to its 1 year specs.

Cheers,
Greg

----- Original Message -----
From: "Brooke Clarke" brooke@pacific.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Saturday, August 06, 2011 6:45 AM
Subject: Re: [volt-nuts] HP 3458A

Hi Greg:

For example on the 10 VDC measurement what are the specs for the two
types of cal?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.End2PartyGovernment.com/

gbusg wrote:

Steve,

Both cal versions verify all functions and ranges of the 3458A, and both
versions provide test data.

However the Standards Lab cal (known as a "Golden calibration) utilizes a
completely different, more sophisticated procedure and methodology,
resulting in significantly lower Measurement Uncertainty for most
measurands
(compared to the STE9000 calibration).

If you want to trend your 3458A at specific measurands (e.g., at 10Vdc,
10k
ohms and 1V 20kHz, etc.), then you will want the Standards Lab cal - this
is
because the STE9000 calibration's Test Uncertainty Ratios are too low for
you to realize enough meaningful confidence and repeatability of the data
for the purpose of trending specific measurands.

For the same reason you will want the Standards Lab cal if you plan to use
Agilent's calibration test report data as correction factors in some
state-of-art process you have.

If neither of these two applications fit you, then the STE9000 calibration
will probably suffice for you.

The more I think about this, I think mostly it's other metrology labs who
need the "Golden calibration" (for at least one of their 3458As) and those
are the folks who already know what they need and how to order it.

-Greg

----- Original Message -----
From: "Steve"steve-krull@cox.net
To: "Discussion of precise voltage measurement"volt-nuts@febo.com
Cc:volt-nuts@febo.com
Sent: Friday, August 05, 2011 2:28 PM
Subject: Re: [volt-nuts] HP 3458A

I haven't sent the meter in for calibration yet. Hoping Santa might bring
that for Christmas. Our local Agilent rep swore the only difference
between
the Agilent $550 calibration per incident and the pricier ones offered is
the amount of paperwork you receive; the actual calibration is to full
specs
for all functions and all ranges. The Agilent web site seems to say the
same
thing, so I'm a bit confused by others saying there's calibration and
there's full calibration. I need to go read the information provided by
Greg
Burnett and then approach Agilent again. When I was in metrology full
time,
all calibrations were to full specs or you had to clearly note any
deviations and get the customer to buy off on them. It was amazing how
many
would accept things I wouldn't accept for my home lab!

Steve


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Calibration, is only testing and noting the deviation from the standard used. To adjust the meter to this standard is an extra service. And indeed shipping, other air humidity, other temperature ect and time will change this. I have a guildline cabinet. The papers it was calibrated every year in the 79's and 80's. I have the reports ( and some copy of the bills) because it came with all documentation . I got it as a gift from an internal calibration lab of a very big company. I was lookimg on the net to search info and I wrote some firms like guildline to find out more about these wonderfull cabinets. Then I heard that calibrating such a cabinet ( that was, if I lived in the USA, but over here it must be allso posdible I guess) was still possible and transportation alone was 750 dollar) Fred PA4TIM Op 7 aug. 2011 om 02:13 heeft WB6BNQ <wb6bnq@cox.net> het volgende geschreven: > This response illustrates the absurdity in the whole process. The so-called > "Golden Calibration" is only of value in shifting from a fixed standard to an > instrument that can measure "in-between" values to a high degree if done within > minutes of the set up. It is called the transfer ratio and is the 24 hour > specification. > > The absurdity to think your going spend additional money for the "Golden" part > and get something is total BS. Even the STE/9000 is the value at the time of > calibration. Bouncing around in shipping could certainly invalidate that value. > > Such antics could best be described as the cat chasing his tail. > > Bill....WB6BNQ > > > gbusg wrote: > >> Hi Brooke, >> >> MU at 10Vdc for the "Golden Calibration" is in the range +/- 0.2 ppm, IIRC. >> >> MU at 10Vdc for the "STE/9000 Calibration" is in the range +/- 3.6 ppm, >> IIRC. >> >> The "Golden Calibration" verifies the 3458A to its 24 hour specs. >> >> The "STE/9000 Calibration" verifies the 3458A to its 1 year specs. >> >> Cheers, >> Greg >> >> ----- Original Message ----- >> From: "Brooke Clarke" <brooke@pacific.net> >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> Sent: Saturday, August 06, 2011 6:45 AM >> Subject: Re: [volt-nuts] HP 3458A >> >> Hi Greg: >> >> For example on the 10 VDC measurement what are the specs for the two >> types of cal? >> >> Have Fun, >> >> Brooke Clarke >> http://www.PRC68.com >> http://www.End2PartyGovernment.com/ >> >> gbusg wrote: >>> Steve, >>> >>> Both cal versions verify all functions and ranges of the 3458A, and both >>> versions provide test data. >>> >>> However the Standards Lab cal (known as a "Golden calibration) utilizes a >>> completely different, more sophisticated procedure and methodology, >>> resulting in significantly lower Measurement Uncertainty for most >>> measurands >>> (compared to the STE9000 calibration). >>> >>> If you want to trend your 3458A at specific measurands (e.g., at 10Vdc, >>> 10k >>> ohms and 1V 20kHz, etc.), then you will want the Standards Lab cal - this >>> is >>> because the STE9000 calibration's Test Uncertainty Ratios are too low for >>> you to realize enough meaningful confidence and repeatability of the data >>> for the purpose of trending specific measurands. >>> >>> For the same reason you will want the Standards Lab cal if you plan to use >>> Agilent's calibration test report data as correction factors in some >>> state-of-art process you have. >>> >>> If neither of these two applications fit you, then the STE9000 calibration >>> will probably suffice for you. >>> >>> The more I think about this, I think mostly it's other metrology labs who >>> need the "Golden calibration" (for at least one of their 3458As) and those >>> are the folks who already know what they need and how to order it. >>> >>> -Greg >>> >>> ----- Original Message ----- >>> From: "Steve"<steve-krull@cox.net> >>> To: "Discussion of precise voltage measurement"<volt-nuts@febo.com> >>> Cc:<volt-nuts@febo.com> >>> Sent: Friday, August 05, 2011 2:28 PM >>> Subject: Re: [volt-nuts] HP 3458A >>> >>> >>> I haven't sent the meter in for calibration yet. Hoping Santa might bring >>> that for Christmas. Our local Agilent rep swore the only difference >>> between >>> the Agilent $550 calibration per incident and the pricier ones offered is >>> the amount of paperwork you receive; the actual calibration is to full >>> specs >>> for all functions and all ranges. The Agilent web site seems to say the >>> same >>> thing, so I'm a bit confused by others saying there's calibration and >>> there's full calibration. I need to go read the information provided by >>> Greg >>> Burnett and then approach Agilent again. When I was in metrology full >>> time, >>> all calibrations were to full specs or you had to clearly note any >>> deviations and get the customer to buy off on them. It was amazing how >>> many >>> would accept things I wouldn't accept for my home lab! >>> >>> Steve >>> >>> >>> >>> _______________________________________________ >>> volt-nuts mailing list -- volt-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>> and follow the instructions there. >>> >>> >>> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
PK
Poul-Henning Kamp
Sun, Aug 7, 2011 5:41 AM

In message 4E3DD8AF.FF938E47@cox.net, WB6BNQ writes:

This response illustrates the absurdity in the whole process.  The so-called
"Golden Calibration" is only of value in shifting from a fixed standard to an
instrument that can measure "in-between" values to a high degree if done within
minutes of the set up.

Bill, think about it for a moment, isn't that exactly what you would want
your secondary calibration lab to do ?

If you use the 3458A as bench-instrument, you want it to be stable for
multiple years so you save on calibration cost/cycles.

If you use the 3458A to compare your josephson standard to the customers
Fluke 73x, you want rock steady short term stability, but the absolute
precision doesn't really matter, as long as it is stable.

Just because you don't need it, or see the point, doesn't make it "BS".

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <4E3DD8AF.FF938E47@cox.net>, WB6BNQ writes: >This response illustrates the absurdity in the whole process. The so-called >"Golden Calibration" is only of value in shifting from a fixed standard to an >instrument that can measure "in-between" values to a high degree if done within >minutes of the set up. Bill, think about it for a moment, isn't that exactly what you would want your secondary calibration lab to do ? If you use the 3458A as bench-instrument, you want it to be stable for multiple years so you save on calibration cost/cycles. If you use the 3458A to compare your josephson standard to the customers Fluke 73x, you want rock steady short term stability, but the absolute precision doesn't really matter, as long as it is stable. Just because _you_ don't need it, or see the point, doesn't make it "BS". -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
FS
Fred Schneider
Sun, Aug 7, 2011 5:55 AM

I agree, i have my calibration stuff, just for hobby. I tried to get everything as good as possible. I'm rather sure I can get accurate to 10 uV. Probably a bit better. That is not perfect but more then decent to get my collection of old multimeters and scopes accurate again. I bought a Calibrated new 6,5 Keithley, measured my weston cells immeiate after receiving it and a few hours wrming up. That was in line with the known specs and history from the company I got it from. I checked that with more meters and agains some calibrators and I use the 10 uV limit as worst case.  That is my golden standard. I use a solartron 7061as reference. Before calibration I calibrate my solartron using the weston cells and after that the solartron to check the calibrators and dividers. I have to try to do the best as I can but it is a hobby, I do sometimes a calibration for a friend but free of charge, so no garantees ;-)

Fred PA4TIM

Op 7 aug. 2011 om 07:41 heeft "Poul-Henning Kamp" phk@phk.freebsd.dk het volgende geschreven:

In message 4E3DD8AF.FF938E47@cox.net, WB6BNQ writes:

This response illustrates the absurdity in the whole process.  The so-called
"Golden Calibration" is only of value in shifting from a fixed standard to an
instrument that can measure "in-between" values to a high degree if done within
minutes of the set up.

Bill, think about it for a moment, isn't that exactly what you would want
your secondary calibration lab to do ?

If you use the 3458A as bench-instrument, you want it to be stable for
multiple years so you save on calibration cost/cycles.

If you use the 3458A to compare your josephson standard to the customers
Fluke 73x, you want rock steady short term stability, but the absolute
precision doesn't really matter, as long as it is stable.

Just because you don't need it, or see the point, doesn't make it "BS".

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I agree, i have my calibration stuff, just for hobby. I tried to get everything as good as possible. I'm rather sure I can get accurate to 10 uV. Probably a bit better. That is not perfect but more then decent to get my collection of old multimeters and scopes accurate again. I bought a Calibrated new 6,5 Keithley, measured my weston cells immeiate after receiving it and a few hours wrming up. That was in line with the known specs and history from the company I got it from. I checked that with more meters and agains some calibrators and I use the 10 uV limit as worst case. That is my golden standard. I use a solartron 7061as reference. Before calibration I calibrate my solartron using the weston cells and after that the solartron to check the calibrators and dividers. I have to try to do the best as I can but it is a hobby, I do sometimes a calibration for a friend but free of charge, so no garantees ;-) Fred PA4TIM Op 7 aug. 2011 om 07:41 heeft "Poul-Henning Kamp" <phk@phk.freebsd.dk> het volgende geschreven: > In message <4E3DD8AF.FF938E47@cox.net>, WB6BNQ writes: > >> This response illustrates the absurdity in the whole process. The so-called >> "Golden Calibration" is only of value in shifting from a fixed standard to an >> instrument that can measure "in-between" values to a high degree if done within >> minutes of the set up. > > Bill, think about it for a moment, isn't that exactly what you would want > your secondary calibration lab to do ? > > If you use the 3458A as bench-instrument, you want it to be stable for > multiple years so you save on calibration cost/cycles. > > If you use the 3458A to compare your josephson standard to the customers > Fluke 73x, you want rock steady short term stability, but the absolute > precision doesn't really matter, as long as it is stable. > > Just because _you_ don't need it, or see the point, doesn't make it "BS". > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
S
Steve
Sun, Aug 7, 2011 6:45 PM

Greg, Marv, Bill, et al,

This has been a very enlightening thread and I appreciate everyone's input. From a hobbyist perspective the STE9000 cal is all I'll need unless a contract should come along that would demand MU better than that.

It's too bad the Agilent cal services description isn't as clear as what's available on this list! Of course once the explanations supplied here are applied to the Agilent descriptions they become much clearer.

Best regards,

Steve

On Aug 6, 2011, at 5:14 AM, "gbusg" gbusg@comcast.net wrote:

Steve,

Both cal versions verify all functions and ranges of the 3458A, and both
versions provide test data.

However the Standards Lab cal (known as a "Golden calibration) utilizes a
completely different, more sophisticated procedure and methodology,
resulting in significantly lower Measurement Uncertainty for most measurands
(compared to the STE9000 calibration).

If you want to trend your 3458A at specific measurands (e.g., at 10Vdc, 10k
ohms and 1V 20kHz, etc.), then you will want the Standards Lab cal - this is
because the STE9000 calibration's Test Uncertainty Ratios are too low for
you to realize enough meaningful confidence and repeatability of the data
for the purpose of trending specific measurands.

For the same reason you will want the Standards Lab cal if you plan to use
Agilent's calibration test report data as correction factors in some
state-of-art process you have.

If neither of these two applications fit you, then the STE9000 calibration
will probably suffice for you.

The more I think about this, I think mostly it's other metrology labs who
need the "Golden calibration" (for at least one of their 3458As) and those
are the folks who already know what they need and how to order it.

-Greg

----- Original Message -----
From: "Steve" steve-krull@cox.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Cc: volt-nuts@febo.com
Sent: Friday, August 05, 2011 2:28 PM
Subject: Re: [volt-nuts] HP 3458A

I haven't sent the meter in for calibration yet. Hoping Santa might bring
that for Christmas. Our local Agilent rep swore the only difference between
the Agilent $550 calibration per incident and the pricier ones offered is
the amount of paperwork you receive; the actual calibration is to full specs
for all functions and all ranges. The Agilent web site seems to say the same
thing, so I'm a bit confused by others saying there's calibration and
there's full calibration. I need to go read the information provided by Greg
Burnett and then approach Agilent again. When I was in metrology full time,
all calibrations were to full specs or you had to clearly note any
deviations and get the customer to buy off on them. It was amazing how many
would accept things I wouldn't accept for my home lab!

Steve


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Greg, Marv, Bill, et al, This has been a very enlightening thread and I appreciate everyone's input. From a hobbyist perspective the STE9000 cal is all I'll need unless a contract should come along that would demand MU better than that. It's too bad the Agilent cal services description isn't as clear as what's available on this list! Of course once the explanations supplied here are applied to the Agilent descriptions they become much clearer. Best regards, Steve On Aug 6, 2011, at 5:14 AM, "gbusg" <gbusg@comcast.net> wrote: > Steve, > > Both cal versions verify all functions and ranges of the 3458A, and both > versions provide test data. > > However the Standards Lab cal (known as a "Golden calibration) utilizes a > completely different, more sophisticated procedure and methodology, > resulting in significantly lower Measurement Uncertainty for most measurands > (compared to the STE9000 calibration). > > If you want to trend your 3458A at specific measurands (e.g., at 10Vdc, 10k > ohms and 1V 20kHz, etc.), then you will want the Standards Lab cal - this is > because the STE9000 calibration's Test Uncertainty Ratios are too low for > you to realize enough meaningful confidence and repeatability of the data > for the purpose of trending specific measurands. > > For the same reason you will want the Standards Lab cal if you plan to use > Agilent's calibration test report data as correction factors in some > state-of-art process you have. > > If neither of these two applications fit you, then the STE9000 calibration > will probably suffice for you. > > The more I think about this, I think mostly it's other metrology labs who > need the "Golden calibration" (for at least one of their 3458As) and those > are the folks who already know what they need and how to order it. > > -Greg > > ----- Original Message ----- > From: "Steve" <steve-krull@cox.net> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Cc: <volt-nuts@febo.com> > Sent: Friday, August 05, 2011 2:28 PM > Subject: Re: [volt-nuts] HP 3458A > > > I haven't sent the meter in for calibration yet. Hoping Santa might bring > that for Christmas. Our local Agilent rep swore the only difference between > the Agilent $550 calibration per incident and the pricier ones offered is > the amount of paperwork you receive; the actual calibration is to full specs > for all functions and all ranges. The Agilent web site seems to say the same > thing, so I'm a bit confused by others saying there's calibration and > there's full calibration. I need to go read the information provided by Greg > Burnett and then approach Agilent again. When I was in metrology full time, > all calibrations were to full specs or you had to clearly note any > deviations and get the customer to buy off on them. It was amazing how many > would accept things I wouldn't accept for my home lab! > > Steve > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
G
gbusg
Sun, Aug 7, 2011 7:28 PM

Your argument would be valid where Type A Uncertainty and/or drift are the
dominant contributors to a specification. However that's not the case for
the HP/Agilent 3458A, where Type B Uncertainty is the dominant contributor.
Over the past 21 years metrology labs and National Metrology Institutes have
successfully used this fact to their advantage by using 'Golden' 3458As in
order to reduce the propagation of Type B Uncertainty in their state-of-art
applications.

This is the same technique used by the same labs to reduce propagation of
Type B Uncertainty in their state-of-art processes for other devices like
attenuators, thermal converters and power sensors, etc.

That this opportunity exists with the 3458A speaks very highly of it.

Keep in mind that, for all such 'golden' processes, it's necessary to use
the standards labs' measurement data (of your 'golden' instrument) as
correction factors in your processes.

If you're not actively involved (with state-of-art processes requiring such
reduction of the propagation of Type B Uncertainty) - and if you're not
actively using a 3458A in a state-of-art application like this - then don't
worry about it.

http://lvldstdslabagilent.blogspot.com/2005/01/faq-what-3458a-calibrations-are.html

http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

-Greg

----- Original Message -----
From: "WB6BNQ" wb6bnq@cox.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Saturday, August 06, 2011 6:13 PM
Subject: Re: [volt-nuts] HP 3458A

This response illustrates the absurdity in the whole process.  The so-called
"Golden Calibration" is only of value in shifting from a fixed standard to
an
instrument that can measure "in-between" values to a high degree if done
within
minutes of the set up.  It is called the transfer ratio and is the 24 hour
specification.

The absurdity to think your going spend additional money for the "Golden"
part
and get something is total BS.  Even the STE/9000 is the value at the time
of
calibration.  Bouncing around in shipping could certainly invalidate that
value.

Such antics could best be described as the cat chasing his tail.

Bill....WB6BNQ

Your argument would be valid where Type A Uncertainty and/or drift are the dominant contributors to a specification. However that's not the case for the HP/Agilent 3458A, where Type B Uncertainty is the dominant contributor. Over the past 21 years metrology labs and National Metrology Institutes have successfully used this fact to their advantage by using 'Golden' 3458As in order to reduce the propagation of Type B Uncertainty in their state-of-art applications. This is the same technique used by the same labs to reduce propagation of Type B Uncertainty in their state-of-art processes for other devices like attenuators, thermal converters and power sensors, etc. That this opportunity exists with the 3458A speaks very highly of it. Keep in mind that, for all such 'golden' processes, it's necessary to use the standards labs' measurement data (of your 'golden' instrument) as correction factors in your processes. If you're not actively involved (with state-of-art processes requiring such reduction of the propagation of Type B Uncertainty) - and if you're not actively using a 3458A in a state-of-art application like this - then don't worry about it. http://lvldstdslabagilent.blogspot.com/2005/01/faq-what-3458a-calibrations-are.html http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf -Greg ----- Original Message ----- From: "WB6BNQ" <wb6bnq@cox.net> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Saturday, August 06, 2011 6:13 PM Subject: Re: [volt-nuts] HP 3458A This response illustrates the absurdity in the whole process. The so-called "Golden Calibration" is only of value in shifting from a fixed standard to an instrument that can measure "in-between" values to a high degree if done within minutes of the set up. It is called the transfer ratio and is the 24 hour specification. The absurdity to think your going spend additional money for the "Golden" part and get something is total BS. Even the STE/9000 is the value at the time of calibration. Bouncing around in shipping could certainly invalidate that value. Such antics could best be described as the cat chasing his tail. Bill....WB6BNQ
W
WB6BNQ
Mon, Aug 8, 2011 6:47 AM

Hi Poul,

I guess you missed my point.

Unless you are a business that relies on a traceable process, such as a real Second
Level Lab, it makes no sense for the hobbyist to spend the money to go beyond the
basic calibration service.  To the hobbyist, for all intent and purpose, the basic
calibration would exceed any reasonable measurement requirement the hobbyist could
come up with.

Unless a hobbyist has a lot of money, the average hobbyist person is not going to be
able to provide a Lab area that would be controllable to within one degree
centigrade.  That is the level needed to make real use of the 24 hour specification
and that specification is using the 3458A as a transfer standard.

The only thing a real Primary Lab, with the capability, could do that would be worth
the value would be to use the JJ array to verify the actual linearity of the A/D of
that specific meter.  That would also evaluate the meter’s internal reference to an
absolute standard.  In all truth, I doubt that is what is being classed as a
so-called “Golden” calibration because the money is not high enough to cover the
time and cost required for a full verification of the meter’s A/D linearity and the
meter’s absolute stability.  HP has already done such work in the verification of
their production line of the 3458A to prove its worth.  That effort, assuming you
accept it, has shown the validity of the circuit design, repeatability and stability
in production units.

The real problem is the average person getting caught up the numbers game where they
have no way in hell proving it one way or another.  They also have no reason for
such so-called extreme accuracy and resolution.  Sure it is a damn good meter and if
you can afford one it is a very nice meter to have.  A hobbyist who owns one should
consider himself as being lucky and that it only cost around $500 to get it
calibrated if you have the means.  They should be especially thankful that they can
have the calibration from a, hopefully, trustworthy source, such as the very maker
of the instrument, at such a reasonable price.

For the hobbyist, making smart use of the instrument with basic calibration and
understanding how to use it properly would be way more profitable then getting
caught up in the numbers game to an insane level.

My two cents.

Bill....WB6BNQ

Poul-Henning Kamp wrote:

In message 4E3DD8AF.FF938E47@cox.net, WB6BNQ writes:

This response illustrates the absurdity in the whole process.  The so-called
"Golden Calibration" is only of value in shifting from a fixed standard to an
instrument that can measure "in-between" values to a high degree if done within
minutes of the set up.

Bill, think about it for a moment, isn't that exactly what you would want
your secondary calibration lab to do ?

If you use the 3458A as bench-instrument, you want it to be stable for
multiple years so you save on calibration cost/cycles.

If you use the 3458A to compare your josephson standard to the customers
Fluke 73x, you want rock steady short term stability, but the absolute
precision doesn't really matter, as long as it is stable.

Just because you don't need it, or see the point, doesn't make it "BS".

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Poul, I guess you missed my point. Unless you are a business that relies on a traceable process, such as a real Second Level Lab, it makes no sense for the hobbyist to spend the money to go beyond the basic calibration service. To the hobbyist, for all intent and purpose, the basic calibration would exceed any reasonable measurement requirement the hobbyist could come up with. Unless a hobbyist has a lot of money, the average hobbyist person is not going to be able to provide a Lab area that would be controllable to within one degree centigrade. That is the level needed to make real use of the 24 hour specification and that specification is using the 3458A as a transfer standard. The only thing a real Primary Lab, with the capability, could do that would be worth the value would be to use the JJ array to verify the actual linearity of the A/D of that specific meter. That would also evaluate the meter’s internal reference to an absolute standard. In all truth, I doubt that is what is being classed as a so-called “Golden” calibration because the money is not high enough to cover the time and cost required for a full verification of the meter’s A/D linearity and the meter’s absolute stability. HP has already done such work in the verification of their production line of the 3458A to prove its worth. That effort, assuming you accept it, has shown the validity of the circuit design, repeatability and stability in production units. The real problem is the average person getting caught up the numbers game where they have no way in hell proving it one way or another. They also have no reason for such so-called extreme accuracy and resolution. Sure it is a damn good meter and if you can afford one it is a very nice meter to have. A hobbyist who owns one should consider himself as being lucky and that it only cost around $500 to get it calibrated if you have the means. They should be especially thankful that they can have the calibration from a, hopefully, trustworthy source, such as the very maker of the instrument, at such a reasonable price. For the hobbyist, making smart use of the instrument with basic calibration and understanding how to use it properly would be way more profitable then getting caught up in the numbers game to an insane level. My two cents. Bill....WB6BNQ Poul-Henning Kamp wrote: > In message <4E3DD8AF.FF938E47@cox.net>, WB6BNQ writes: > > >This response illustrates the absurdity in the whole process. The so-called > >"Golden Calibration" is only of value in shifting from a fixed standard to an > >instrument that can measure "in-between" values to a high degree if done within > >minutes of the set up. > > Bill, think about it for a moment, isn't that exactly what you would want > your secondary calibration lab to do ? > > If you use the 3458A as bench-instrument, you want it to be stable for > multiple years so you save on calibration cost/cycles. > > If you use the 3458A to compare your josephson standard to the customers > Fluke 73x, you want rock steady short term stability, but the absolute > precision doesn't really matter, as long as it is stable. > > Just because _you_ don't need it, or see the point, doesn't make it "BS". > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
W
WB6BNQ
Mon, Aug 8, 2011 7:22 AM

Greg,

While I can appreciate your enthusiasm, I think you are caught up in the numbers game.
Unless I am mistaken, most people on this list are dealing with their personal
hobbies.  As such their efforts are done on a limited budget and playing with the big
boys gets expensive.

So my comments are coming from that perspective.  Also, my comments are coming from
what is reasonable in the real world as it relates to the hobbyist or amateur
experimenter.

And NO Greg, I have no idea what you mean by type A or B.  How about explaining that ?

Your first link goes to someone’s blog whose data amounts to heresy.  Where is the
information from a valid HP site ?  If you actually go to a HP site on the 3458A, you
will see that the additional Calibration costs are referring to additional verification
paperwork.  If there is such a thing as a “Golden” anything, then it needs to properly
explained; not spoke of as though you got to have the secret password and decoder ring
to get any information.  Those type of comments are generally born out of
misunderstandings and little substance.

Your second link references a HP Journal about the meter from over 20 years ago.  How
does that support or address your comments ?  If your trying to point to the
complexity, then see my response to Poul.

Bill....WB6BNQ

gbusg wrote:

Your argument would be valid where Type A Uncertainty and/or drift are the
dominant contributors to a specification. However that's not the case for
the HP/Agilent 3458A, where Type B Uncertainty is the dominant contributor.
Over the past 21 years metrology labs and National Metrology Institutes have
successfully used this fact to their advantage by using 'Golden' 3458As in
order to reduce the propagation of Type B Uncertainty in their state-of-art
applications.

This is the same technique used by the same labs to reduce propagation of
Type B Uncertainty in their state-of-art processes for other devices like
attenuators, thermal converters and power sensors, etc.

That this opportunity exists with the 3458A speaks very highly of it.

Keep in mind that, for all such 'golden' processes, it's necessary to use
the standards labs' measurement data (of your 'golden' instrument) as
correction factors in your processes.

If you're not actively involved (with state-of-art processes requiring such
reduction of the propagation of Type B Uncertainty) - and if you're not
actively using a 3458A in a state-of-art application like this - then don't
worry about it.

http://lvldstdslabagilent.blogspot.com/2005/01/faq-what-3458a-calibrations-are.html

http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

-Greg

----- Original Message -----
From: "WB6BNQ" wb6bnq@cox.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Saturday, August 06, 2011 6:13 PM
Subject: Re: [volt-nuts] HP 3458A

This response illustrates the absurdity in the whole process.  The so-called
"Golden Calibration" is only of value in shifting from a fixed standard to
an
instrument that can measure "in-between" values to a high degree if done
within
minutes of the set up.  It is called the transfer ratio and is the 24 hour
specification.

The absurdity to think your going spend additional money for the "Golden"
part
and get something is total BS.  Even the STE/9000 is the value at the time
of
calibration.  Bouncing around in shipping could certainly invalidate that
value.

Such antics could best be described as the cat chasing his tail.

Bill....WB6BNQ


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Greg, While I can appreciate your enthusiasm, I think you are caught up in the numbers game. Unless I am mistaken, most people on this list are dealing with their personal hobbies. As such their efforts are done on a limited budget and playing with the big boys gets expensive. So my comments are coming from that perspective. Also, my comments are coming from what is reasonable in the real world as it relates to the hobbyist or amateur experimenter. And NO Greg, I have no idea what you mean by type A or B. How about explaining that ? Your first link goes to someone’s blog whose data amounts to heresy. Where is the information from a valid HP site ? If you actually go to a HP site on the 3458A, you will see that the additional Calibration costs are referring to additional verification paperwork. If there is such a thing as a “Golden” anything, then it needs to properly explained; not spoke of as though you got to have the secret password and decoder ring to get any information. Those type of comments are generally born out of misunderstandings and little substance. Your second link references a HP Journal about the meter from over 20 years ago. How does that support or address your comments ? If your trying to point to the complexity, then see my response to Poul. Bill....WB6BNQ gbusg wrote: > Your argument would be valid where Type A Uncertainty and/or drift are the > dominant contributors to a specification. However that's not the case for > the HP/Agilent 3458A, where Type B Uncertainty is the dominant contributor. > Over the past 21 years metrology labs and National Metrology Institutes have > successfully used this fact to their advantage by using 'Golden' 3458As in > order to reduce the propagation of Type B Uncertainty in their state-of-art > applications. > > This is the same technique used by the same labs to reduce propagation of > Type B Uncertainty in their state-of-art processes for other devices like > attenuators, thermal converters and power sensors, etc. > > That this opportunity exists with the 3458A speaks very highly of it. > > Keep in mind that, for all such 'golden' processes, it's necessary to use > the standards labs' measurement data (of your 'golden' instrument) as > correction factors in your processes. > > If you're not actively involved (with state-of-art processes requiring such > reduction of the propagation of Type B Uncertainty) - and if you're not > actively using a 3458A in a state-of-art application like this - then don't > worry about it. > > http://lvldstdslabagilent.blogspot.com/2005/01/faq-what-3458a-calibrations-are.html > > http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf > > -Greg > > ----- Original Message ----- > From: "WB6BNQ" <wb6bnq@cox.net> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Sent: Saturday, August 06, 2011 6:13 PM > Subject: Re: [volt-nuts] HP 3458A > > This response illustrates the absurdity in the whole process. The so-called > "Golden Calibration" is only of value in shifting from a fixed standard to > an > instrument that can measure "in-between" values to a high degree if done > within > minutes of the set up. It is called the transfer ratio and is the 24 hour > specification. > > The absurdity to think your going spend additional money for the "Golden" > part > and get something is total BS. Even the STE/9000 is the value at the time > of > calibration. Bouncing around in shipping could certainly invalidate that > value. > > Such antics could best be described as the cat chasing his tail. > > Bill....WB6BNQ > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
G
gbusg
Mon, Aug 8, 2011 10:44 PM

Bill wrote:

Unless I am mistaken, most people on this list are dealing with their
personal
hobbies.

Yes, Bill, your last few posts have persuaded me that volt-nuts is likely
not the right forum for this kind of esoteric information. (I'll remember to
qualify that the next time being tempted to jump-in with information.) :)

And NO Greg, I have no idea what you mean by type A or B.  How about
explaining that ?

Again, apparently we're beyond the scope of volt-nuts interest, but since
you asked (and to close-out this topic), see:

Guide to the Expression of Uncertainty in Measurement
http://www.bipm.org/utils/common/documents/jcgm/JCGM_100_2008_E.pdf

NIST Reference
http://physics.nist.gov/cuu/Uncertainty/basic.html

UKAS Lab-12
http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/LAB12.PDF

-Greg

Bill wrote: >Unless I am mistaken, most people on this list are dealing with their >personal >hobbies. Yes, Bill, your last few posts have persuaded me that volt-nuts is likely not the right forum for this kind of esoteric information. (I'll remember to qualify that the next time being tempted to jump-in with information.) :) >And NO Greg, I have no idea what you mean by type A or B. How about >explaining that ? Again, apparently we're beyond the scope of volt-nuts interest, but since you asked (and to close-out this topic), see: Guide to the Expression of Uncertainty in Measurement http://www.bipm.org/utils/common/documents/jcgm/JCGM_100_2008_E.pdf NIST Reference http://physics.nist.gov/cuu/Uncertainty/basic.html UKAS Lab-12 http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/LAB12.PDF -Greg
CH
Chuck Harris
Mon, Aug 8, 2011 11:06 PM

Volt-nuts is to voltage measurement and standards as time-nuts is to
time measurement and standards.  It should include everyone with an
interest in maintaining voltage (current/resistance) standards.

I, for one, would like to learn about making JJ voltage standards in
a volt-nut setting.

-Chuck Harris

gbusg wrote:

Bill wrote:

Unless I am mistaken, most people on this list are dealing with their
personal
hobbies.

Yes, Bill, your last few posts have persuaded me that volt-nuts is likely
not the right forum for this kind of esoteric information. (I'll remember to
qualify that the next time being tempted to jump-in with information.) :)

And NO Greg, I have no idea what you mean by type A or B.  How about
explaining that ?

Again, apparently we're beyond the scope of volt-nuts interest, but since
you asked (and to close-out this topic), see:

Guide to the Expression of Uncertainty in Measurement
http://www.bipm.org/utils/common/documents/jcgm/JCGM_100_2008_E.pdf

NIST Reference
http://physics.nist.gov/cuu/Uncertainty/basic.html

UKAS Lab-12
http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/LAB12.PDF

-Greg

Volt-nuts is to voltage measurement and standards as time-nuts is to time measurement and standards. It should include everyone with an interest in maintaining voltage (current/resistance) standards. I, for one, would like to learn about making JJ voltage standards in a volt-nut setting. -Chuck Harris gbusg wrote: > Bill wrote: > >> Unless I am mistaken, most people on this list are dealing with their >> personal >> hobbies. > Yes, Bill, your last few posts have persuaded me that volt-nuts is likely > not the right forum for this kind of esoteric information. (I'll remember to > qualify that the next time being tempted to jump-in with information.) :) > >> And NO Greg, I have no idea what you mean by type A or B. How about >> explaining that ? > Again, apparently we're beyond the scope of volt-nuts interest, but since > you asked (and to close-out this topic), see: > > Guide to the Expression of Uncertainty in Measurement > http://www.bipm.org/utils/common/documents/jcgm/JCGM_100_2008_E.pdf > > NIST Reference > http://physics.nist.gov/cuu/Uncertainty/basic.html > > UKAS Lab-12 > http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/LAB12.PDF > > -Greg
FS
Fred Schneider
Mon, Aug 8, 2011 11:15 PM

Chuck,

Would that be possible for an amateur to make. Count me in ;-) but I read a paper about them, if i remember well,  Fluke in germany and the USA must have one. It sounded rather big and very expensive.

Fred PA4TIM

Op 9 aug. 2011 om 01:06 heeft Chuck Harris cfharris@erols.com het volgende geschreven:

Volt-nuts is to voltage measurement and standards as time-nuts is to
time measurement and standards.  It should include everyone with an
interest in maintaining voltage (current/resistance) standards.

I, for one, would like to learn about making JJ voltage standards in
a volt-nut setting.

-Chuck Harris

gbusg wrote:

Bill wrote:

Unless I am mistaken, most people on this list are dealing with their
personal
hobbies.

Yes, Bill, your last few posts have persuaded me that volt-nuts is likely
not the right forum for this kind of esoteric information. (I'll remember to
qualify that the next time being tempted to jump-in with information.) :)

And NO Greg, I have no idea what you mean by type A or B.  How about
explaining that ?

Again, apparently we're beyond the scope of volt-nuts interest, but since
you asked (and to close-out this topic), see:

Guide to the Expression of Uncertainty in Measurement
http://www.bipm.org/utils/common/documents/jcgm/JCGM_100_2008_E.pdf

NIST Reference
http://physics.nist.gov/cuu/Uncertainty/basic.html

UKAS Lab-12
http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/LAB12.PDF

-Greg


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Chuck, Would that be possible for an amateur to make. Count me in ;-) but I read a paper about them, if i remember well, Fluke in germany and the USA must have one. It sounded rather big and very expensive. Fred PA4TIM Op 9 aug. 2011 om 01:06 heeft Chuck Harris <cfharris@erols.com> het volgende geschreven: > Volt-nuts is to voltage measurement and standards as time-nuts is to > time measurement and standards. It should include everyone with an > interest in maintaining voltage (current/resistance) standards. > > I, for one, would like to learn about making JJ voltage standards in > a volt-nut setting. > > -Chuck Harris > > gbusg wrote: >> Bill wrote: >> >>> Unless I am mistaken, most people on this list are dealing with their >>> personal >>> hobbies. >> Yes, Bill, your last few posts have persuaded me that volt-nuts is likely >> not the right forum for this kind of esoteric information. (I'll remember to >> qualify that the next time being tempted to jump-in with information.) :) >> >>> And NO Greg, I have no idea what you mean by type A or B. How about >>> explaining that ? >> Again, apparently we're beyond the scope of volt-nuts interest, but since >> you asked (and to close-out this topic), see: >> >> Guide to the Expression of Uncertainty in Measurement >> http://www.bipm.org/utils/common/documents/jcgm/JCGM_100_2008_E.pdf >> >> NIST Reference >> http://physics.nist.gov/cuu/Uncertainty/basic.html >> >> UKAS Lab-12 >> http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/LAB12.PDF >> >> -Greg > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.