Problems and solution when using a Rigol DDS signal generator for counter testing

EK
Erik Kaashoek
Fri, Aug 19, 2022 8:32 AM

During the tests of a frequency counter using a Rigol DG812 signal
generator locked to an external 10MHz clock it was observed that the
output frequency of the DG812 was not stable
This can clearly be seen in [1], blue trace, where the counter was
warming up and the frequency of the frequency variations changed over time.
To test if the external locking was related to the instability the
external lock was disabled which lead to huge instability. [2] Pink trace
As can be seen in the pink trace the huge variations disappeared when
the coax cable with the 10MHz reference clock was disconnected (at 48
seconds) and appeared again when the cable was reconnected to the clock
input (at 86 seconds), regardless of the Rigol running the whole time on
its internal clock.
This seemed to hint at some interference between the external clock
signal and the internal frequency reference happening when running both
at internal or external clock.
To test this an  external clock signal was applied at the frequency (+/-
0.01 Hz ) of the internal clock.
The instability of the output from the Rigol locked to an external clock
close to is internal frequency disappeared completely ([1], red trace)
and the remaining frequency variations where within the measurement
uncertainty of the used frequency counter.

[1] http://athome.kaashoek.com/time-nuts/Rigol%20locking%202.png
[2] http://athome.kaashoek.com/time-nuts/Rigol%20pulling.png

During the tests of a frequency counter using a Rigol DG812 signal generator locked to an external 10MHz clock it was observed that the output frequency of the DG812 was not stable This can clearly be seen in [1], blue trace, where the counter was warming up and the frequency of the frequency variations changed over time. To test if the external locking was related to the instability the external lock was disabled which lead to huge instability. [2] Pink trace As can be seen in the pink trace the huge variations disappeared when the coax cable with the 10MHz reference clock was disconnected (at 48 seconds) and appeared again when the cable was reconnected to the clock input (at 86 seconds), regardless of the Rigol running the whole time on its internal clock. This seemed to hint at some interference between the external clock signal and the internal frequency reference happening when running both at internal or external clock. To test this an  external clock signal was applied at the frequency (+/- 0.01 Hz ) of the internal clock. The instability of the output from the Rigol locked to an external clock close to is internal frequency disappeared completely ([1], red trace) and the remaining frequency variations where within the measurement uncertainty of the used frequency counter. [1] http://athome.kaashoek.com/time-nuts/Rigol%20locking%202.png [2] http://athome.kaashoek.com/time-nuts/Rigol%20pulling.png
MD
Magnus Danielson
Fri, Aug 19, 2022 12:15 PM

Hi Erik,

On 2022-08-19 10:32, Erik Kaashoek via time-nuts wrote:

During the tests of a frequency counter using a Rigol DG812 signal
generator locked to an external 10MHz clock it was observed that the
output frequency of the DG812 was not stable
This can clearly be seen in [1], blue trace, where the counter was
warming up and the frequency of the frequency variations changed over
time.
To test if the external locking was related to the instability the
external lock was disabled which lead to huge instability. [2] Pink trace
As can be seen in the pink trace the huge variations disappeared when
the coax cable with the 10MHz reference clock was disconnected (at 48
seconds) and appeared again when the cable was reconnected to the
clock input (at 86 seconds), regardless of the Rigol running the whole
time on its internal clock.
This seemed to hint at some interference between the external clock
signal and the internal frequency reference happening when running
both at internal or external clock.
To test this an  external clock signal was applied at the frequency
(+/- 0.01 Hz ) of the internal clock.
The instability of the output from the Rigol locked to an external
clock close to is internal frequency disappeared completely ([1], red
trace) and the remaining frequency variations where within the
measurement uncertainty of the used frequency counter.

[1] http://athome.kaashoek.com/time-nuts/Rigol%20locking%202.png
[2] http://athome.kaashoek.com/time-nuts/Rigol%20pulling.png

A few good experiments.

The details of the lock mechanism can be the culprit here. For instance,
if it is a phase lock, then dead-band mechanisms can explain this.
Similarly can a frequency lock experience similar dead-band mechanisms.

One dead-band mechanism can be the resolution of the steering, causing
one to alternate between nearby values to average near the actual value.

I've run into this before and some designs create such deviations that
show up in both phase and frequency as you look carefully. For many
purposes it may suffice just fine, but for others not so much.

If you set offset too close, it can take time before you see the
adjustment, so test with offsets inbetween to validate that frequency
gone down.

Yes, I've been hit by similar issues before. Any generator will cause
phase-deviations so question is just how small deviations.

Cheers,
Magnus

Hi Erik, On 2022-08-19 10:32, Erik Kaashoek via time-nuts wrote: > During the tests of a frequency counter using a Rigol DG812 signal > generator locked to an external 10MHz clock it was observed that the > output frequency of the DG812 was not stable > This can clearly be seen in [1], blue trace, where the counter was > warming up and the frequency of the frequency variations changed over > time. > To test if the external locking was related to the instability the > external lock was disabled which lead to huge instability. [2] Pink trace > As can be seen in the pink trace the huge variations disappeared when > the coax cable with the 10MHz reference clock was disconnected (at 48 > seconds) and appeared again when the cable was reconnected to the > clock input (at 86 seconds), regardless of the Rigol running the whole > time on its internal clock. > This seemed to hint at some interference between the external clock > signal and the internal frequency reference happening when running > both at internal or external clock. > To test this an  external clock signal was applied at the frequency > (+/- 0.01 Hz ) of the internal clock. > The instability of the output from the Rigol locked to an external > clock close to is internal frequency disappeared completely ([1], red > trace) and the remaining frequency variations where within the > measurement uncertainty of the used frequency counter. > > [1] http://athome.kaashoek.com/time-nuts/Rigol%20locking%202.png > [2] http://athome.kaashoek.com/time-nuts/Rigol%20pulling.png A few good experiments. The details of the lock mechanism can be the culprit here. For instance, if it is a phase lock, then dead-band mechanisms can explain this. Similarly can a frequency lock experience similar dead-band mechanisms. One dead-band mechanism can be the resolution of the steering, causing one to alternate between nearby values to average near the actual value. I've run into this before and some designs create such deviations that show up in both phase and frequency as you look carefully. For many purposes it may suffice just fine, but for others not so much. If you set offset too close, it can take time before you see the adjustment, so test with offsets inbetween to validate that frequency gone down. Yes, I've been hit by similar issues before. Any generator will cause phase-deviations so question is just how small deviations. Cheers, Magnus