Hi Everyone,

Some days back I emailed on the group regarding the jitter in the sampling clock of ADC. I  have tried the phase modulation method for producing the jitter.
I used the HP 8648B Signal generator for this experiment. I modulated the RF wave (at 10 MHz) with noise using the modulating input of the Signal Generator. For noise I used a Noise gen. from General Radio Company. It has three options for frequency. 10K, 100K and 5 M Hz.
I observed the output using an oscilloscope. The output wave form was jittery. BUT when changed the noise from 5M to 100K there was an increase in jitter.  Can any one help me out there, why this is happening? According to my understanding decreasing the bandwidth of noise  results in decrease in jitter as well.
Also I want to calculate the jitter at the output. i.e. the amount of noise I am adding. I think I wont be able to do it over the oscilloscope. Any ideas ?

Regards

Bilal


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Bilal

The 8648B phase modulation input has a bandwidth of 10Hz to about 10kHz.

The resultant phase modulation dependency on the noise generator
settings depends on the noise generator specifications.
What was the noise generator model, or can you provide a circuit diagram
so the effect of the noise generator settings can be deduced?

Bruce

Bilal Amin wrote: > Hi Everyone, > > Some days back I emailed on the group regarding the jitter in the sampling clock of ADC. I have tried the phase modulation method for producing the jitter. > I used the HP 8648B Signal generator for this experiment. I modulated the RF wave (at 10 MHz) with noise using the modulating input of the Signal Generator. For noise I used a Noise gen. from General Radio Company. It has three options for frequency. 10K, 100K and 5 M Hz. > I observed the output using an oscilloscope. The output wave form was jittery. BUT when changed the noise from 5M to 100K there was an increase in jitter. Can any one help me out there, why this is happening? According to my understanding decreasing the bandwidth of noise results in decrease in jitter as well. > Also I want to calculate the jitter at the output. i.e. the amount of noise I am adding. I think I wont be able to do it over the oscilloscope. Any ideas ? > > Regards > > Bilal > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > Bilal The 8648B phase modulation input has a bandwidth of 10Hz to about 10kHz. The resultant phase modulation dependency on the noise generator settings depends on the noise generator specifications. What was the noise generator model, or can you provide a circuit diagram so the effect of the noise generator settings can be deduced? Bruce
BA
Bilal Amin
Mon, Mar 19, 2007 12:03 PM

----- Original Message -----
From: "Dr Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Monday, March 19, 2007 9:25 PM
Subject: Re: [time-nuts] Producing jitter with Phase Modulation

Bilal Amin wrote:

Hi Everyone,

Some days back I emailed on the group regarding the jitter in the
sampling clock of ADC. I  have tried the phase modulation method for
producing the jitter.
I used the HP 8648B Signal generator for this experiment. I modulated the
RF wave (at 10 MHz) with noise using the modulating input of the Signal
Generator. For noise I used a Noise gen. from General Radio Company. It
has three options for frequency. 10K, 100K and 5 M Hz.
I observed the output using an oscilloscope. The output wave form was
jittery. BUT when changed the noise from 5M to 100K there was an increase
in jitter.  Can any one help me out there, why this is happening?
According to my understanding decreasing the bandwidth of noise  results
in decrease in jitter as well.
Also I want to calculate the jitter at the output. i.e. the amount of
noise I am adding. I think I wont be able to do it over the oscilloscope.
Any ideas ?

Regards

Bilal


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Bilal

The 8648B phase modulation input has a bandwidth of 10Hz to about 10kHz.

The resultant phase modulation dependency on the noise generator
settings depends on the noise generator specifications.
What was the noise generator model, or can you provide a circuit diagram
so the effect of the noise generator settings can be deduced?

Bruce


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Hi Bruce,

I am attaching the schematic with this email. It is actually Random Noise
Generator by General Radio Company 1390 B. It is a very old model but I have
a limited budget so I am trying to do the experiment using this. I need to
produce 10 MHz jittery clock and the maximum frequency for this noise gen.
is 5MHz as you can see from schematic.

Regards
Bilal
Bilal

----- Original Message ----- From: "Dr Bruce Griffiths" <bruce.griffiths@xtra.co.nz> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Sent: Monday, March 19, 2007 9:25 PM Subject: Re: [time-nuts] Producing jitter with Phase Modulation > Bilal Amin wrote: >> Hi Everyone, >> >> Some days back I emailed on the group regarding the jitter in the >> sampling clock of ADC. I have tried the phase modulation method for >> producing the jitter. >> I used the HP 8648B Signal generator for this experiment. I modulated the >> RF wave (at 10 MHz) with noise using the modulating input of the Signal >> Generator. For noise I used a Noise gen. from General Radio Company. It >> has three options for frequency. 10K, 100K and 5 M Hz. >> I observed the output using an oscilloscope. The output wave form was >> jittery. BUT when changed the noise from 5M to 100K there was an increase >> in jitter. Can any one help me out there, why this is happening? >> According to my understanding decreasing the bandwidth of noise results >> in decrease in jitter as well. >> Also I want to calculate the jitter at the output. i.e. the amount of >> noise I am adding. I think I wont be able to do it over the oscilloscope. >> Any ideas ? >> >> Regards >> >> Bilal >> _______________________________________________ >> time-nuts mailing list >> time-nuts@febo.com >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> > Bilal > > The 8648B phase modulation input has a bandwidth of 10Hz to about 10kHz. > > The resultant phase modulation dependency on the noise generator > settings depends on the noise generator specifications. > What was the noise generator model, or can you provide a circuit diagram > so the effect of the noise generator settings can be deduced? > > Bruce > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > Hi Bruce, I am attaching the schematic with this email. It is actually Random Noise Generator by General Radio Company 1390 B. It is a very old model but I have a limited budget so I am trying to do the experiment using this. I need to produce 10 MHz jittery clock and the maximum frequency for this noise gen. is 5MHz as you can see from schematic. Regards Bilal Bilal
P
Pete
Mon, Mar 19, 2007 3:42 PM

Bilal,

The GR noise generators were specified for total noise voltage in a
specified bandwidth. With the HP signal generator input limited to 10KHz,
your observation of increasing noise as the GR source bandwidth is reduced
is as expected. You are simply applying more of the GR's output to the HP
signal generator by better matching the source bandwidth to the "load".

An oscilloscope is a difficult tool for the measurement you propose. If
that's what you must use, then get a storage 'scope or an equivalent digital
'scope which will allow you to integrate up the jittering waveform. This is
essentially what folks do to create an "eye" diagram for estimating high
speed data link jitter. The difficulty with this approach is that the
probability observing maximum jitter is quite low & the 'scope waveform
tends to have very low intensity for these "rare" events, so it's tough to
know where the outer edges of the jitter are. The digital "scopes are
superior for this application, but rather expensive.

Regards,
Pete Rawson

Bilal, The GR noise generators were specified for total noise voltage in a specified bandwidth. With the HP signal generator input limited to 10KHz, your observation of increasing noise as the GR source bandwidth is reduced is as expected. You are simply applying more of the GR's output to the HP signal generator by better matching the source bandwidth to the "load". An oscilloscope is a difficult tool for the measurement you propose. If that's what you must use, then get a storage 'scope or an equivalent digital 'scope which will allow you to integrate up the jittering waveform. This is essentially what folks do to create an "eye" diagram for estimating high speed data link jitter. The difficulty with this approach is that the probability observing maximum jitter is quite low & the 'scope waveform tends to have very low intensity for these "rare" events, so it's tough to know where the outer edges of the jitter are. The digital "scopes are superior for this application, but rather expensive. Regards, Pete Rawson
BA
Bilal Amin
Tue, Mar 20, 2007 3:35 AM

Pete,
Thanks for your valuable information. You have explained exactly what is
happening and I have got it finally. I have a digi. oscilloscope with a
jitter measurement option, but it is only for square waveforms.
Unfortunately the Sig. Gen. I am using has no option for square wave. I
think I have to use a comparator to get the square wave form and then
measure jitter with oscilloscope.

Cheers
Bilal
----- Original Message -----
From: "Pete" peterawson@earthlink.net
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Tuesday, March 20, 2007 2:42 AM
Subject: Re: [time-nuts] Producing jitter with Phase Modulation

Bilal,

The GR noise generators were specified for total noise voltage in a
specified bandwidth. With the HP signal generator input limited to 10KHz,
your observation of increasing noise as the GR source bandwidth is reduced
is as expected. You are simply applying more of the GR's output to the HP
signal generator by better matching the source bandwidth to the "load".

An oscilloscope is a difficult tool for the measurement you propose. If
that's what you must use, then get a storage 'scope or an equivalent
digital
'scope which will allow you to integrate up the jittering waveform. This
is
essentially what folks do to create an "eye" diagram for estimating high
speed data link jitter. The difficulty with this approach is that the
probability observing maximum jitter is quite low & the 'scope waveform
tends to have very low intensity for these "rare" events, so it's tough to
know where the outer edges of the jitter are. The digital "scopes are
superior for this application, but rather expensive.

Regards,
Pete Rawson


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Pete, Thanks for your valuable information. You have explained exactly what is happening and I have got it finally. I have a digi. oscilloscope with a jitter measurement option, but it is only for square waveforms. Unfortunately the Sig. Gen. I am using has no option for square wave. I think I have to use a comparator to get the square wave form and then measure jitter with oscilloscope. Cheers Bilal ----- Original Message ----- From: "Pete" <peterawson@earthlink.net> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Sent: Tuesday, March 20, 2007 2:42 AM Subject: Re: [time-nuts] Producing jitter with Phase Modulation > Bilal, > > The GR noise generators were specified for total noise voltage in a > specified bandwidth. With the HP signal generator input limited to 10KHz, > your observation of increasing noise as the GR source bandwidth is reduced > is as expected. You are simply applying more of the GR's output to the HP > signal generator by better matching the source bandwidth to the "load". > > An oscilloscope is a difficult tool for the measurement you propose. If > that's what you must use, then get a storage 'scope or an equivalent > digital > 'scope which will allow you to integrate up the jittering waveform. This > is > essentially what folks do to create an "eye" diagram for estimating high > speed data link jitter. The difficulty with this approach is that the > probability observing maximum jitter is quite low & the 'scope waveform > tends to have very low intensity for these "rare" events, so it's tough to > know where the outer edges of the jitter are. The digital "scopes are > superior for this application, but rather expensive. > > Regards, > Pete Rawson > > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >