time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Re: [time-nuts] GPS Modulation and 10 MHz Delay Lock (Jim Lux)

J
johncroos@aol.com
Wed, Oct 3, 2012 3:40 AM

Hi Jim -

Thanks for the update on the modern GPS receivers. I was aware that the
modern ones
do not have a classical analog tracking loop, much less a bunch of them.
However it is a useful concept for purposes of explanation that you do
not need the 1 pps to
lock up the 10 MHz VCXO - which was my main point.

The Tbolt block diagram in the manual Figure 5-10 shows the 10 MHz VCXO
output going to
the receiver and also to the output. The 1 pps comes from the "cpu and
support" circuit.

While the diagram is clearly simplistic - it implies that the Receiver
circuitry first locks up the
10 MHz oscillator and the the 1 pps is derived from that.  My only
point was that it is possible
and perhaps even better to discipline an oscillator using a code
correlator (however implemented)
rather than steering using the 1 pps. I believe this is why the T bolts
work so well. For one thing the
loop should work better and faster if the input is at the chipping rate
rather than at 1 pps as the
information rate is higher.

On doppler - I believe that since the spread spectrum sidebands are
coherent with the suppressed carrier, their
relationship to it is unchanged by doppler and thus it should be
possible to achieve a code correlation on
a  doppler shifted signal. The recovered carrier would be shifted in
frequency by the doppler but it would still be
recovered - at least by a classical IF correlator.

  • On jamming - maybe so, but the effect of the receiving correlator is
    to spread the energy of a CW interferer
    and concentrate the energy of the signal with the matching PN
    modulation is it not????

good fun anyway!

-73 john k6iql

-----Original Message-----
From: time-nuts-request time-nuts-request@febo.com
To: time-nuts time-nuts@febo.com
Sent: Tue, Oct 2, 2012 9:53 pm
Subject: time-nuts Digest, Vol 99, Issue 19

Send time-nuts mailing list submissions to
time-nuts@febo.com

To subscribe or unsubscribe via the World Wide Web, visit
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
or, via email, send a message with subject or body 'help' to
time-nuts-request@febo.com

You can reach the person managing the list at
time-nuts-owner@febo.com

When replying, please edit your Subject line so it is more specific
than "Re: Contents of time-nuts digest..."

Today's Topics:

  1. Re: GPS Modulation and 10 MHz Delay Lock (Jim Lux)
  2. Re: GPS Jammer (Jim Lux)
  3. Re: GPS Jammer (johncroos@aol.com)
  4. Re: Best counter setting for ADEV? (Jim Lux)
  5. Re: RFX GPSDO - Anybody played with one of these? (Jim Lux)
  6. Centering ocxo (Bill Dailey)
  7. Re: GPS Jammer (Jim Lux)
  8. Re: Google's Spanner uses GPS and atomic time (paul swed)

Message: 1
Date: Tue, 02 Oct 2012 19:30:20 -0700
From: Jim Lux jimlux@earthlink.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] GPS Modulation and 10 MHz Delay Lock
Message-ID: 506BA33C.1010606@earthlink.net
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

On 10/2/12 3:39 PM, johncroos@aol.com wrote:

Hello All -

Here is a link that describes the GPS modulation. You do not need the
1 pps to lock the 10 MHz oscillator to the atomic clock in the

satellites.

http://www.kowoma.de/en/gps/signals.htm

If you look at the block diagram you see PN code modulates the

carrier at

the 1.023 MHz chip rate. This is done by BPSK modulation of the

carrier

with the PN code. It can be done simply with a double balanced mixer.

This spreads the signal with PSK at the chip( i.e. code clock) rate.

Note also the modulo - 2 addition of the data to the code sequence.

This

called code inversion
modulation. After de-spread of the code in the receiver - the signal

is

then simple BPSK and
may be demodulated by a Costas or Squaring Loop to get at the data

message.

The obtain precision frequency needed I believe the T bolt simply

locks

to the chipping rate
using some form of Delay Lock Loop. It is NOT at PLL. There is no need
what ever to
deal with the 1 pps using this method. The internal 10 MHz oscillator

is

controlled by this locking circuit and
is part of the code correlation loop.

That's not quite how it works.. It would work for terrestrial links
where there is no Doppler, but in the GPS case, there is significant
Doppler shift on all the signals.  Since the carrier and the chips are
generated from a common source on the spacecraft (the carrier frequency
is a multiple of the chip rate, in fact), you can recover carrier and
chips at the same time.

But.. most receivers these days don't actually have an analog tracking
loop at all.  They digitize the input signal (1 bit quantizer) at a
rate
that makes the carrier alias down to something convenient (a few
hundred
kHz is typical.. you want it far enough away from zero that Doppler
never makes it go negative).  In the experimental receiver in SCaN
Testbed flying on ISS it's about 39 MHz sample rate.

Once you've got your one bit samples, you do some sort of combined
Doppler/Code phase acquisition (these days, often using an FFT), then
track both together digitally using some form of NCO.  The tracking
loops for all the satellite signals aren't necessarily independent and
might be part of a Kalman filter that estimates all the observables
together.

Finally, from all that, you have an estimate of your local clock offset
and timing offset, and from that you can generate your 1pps, typically
with another NCO (with granularity of your clock rate).  Since it's
unlikely that your clock is EXACTLY an even number of cycles per
second,
at each second, a bit of error accumulates, until you have an whole
cycle's worth leading to the familiar sawtooth error.

That sawtooth error is predictable, of course, so you can generate a
"time error" estimate for each 1pps pulse (or, even, control a variable
delay to line it up).

The important thing is that in modern receivers, nowhere is there a
signal at the GPS carrier frequency, nor is there a signal at the chip
rate.  There is probably a signal (with low precision) at the code
epoch (every millisecond), but it's different for each satellite
signal,
of course.


Message: 2
Date: Tue, 02 Oct 2012 19:32:48 -0700
From: Jim Lux jimlux@earthlink.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] GPS Jammer
Message-ID: 506BA3D0.4090609@earthlink.net
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

On 10/2/12 4:48 PM, Charles P. Steinmetz wrote:

The Power Output is 0.5 Watts and it claims a jamming range of 1-10
Meters.
Anybody think there is something wrong?

I'd expect a much greater range with a 0.5 W jammer.  But note that

0.5

W is the "total output power" -- the "transmit power" is only 10 dBm
(0.01 W).  Whatever those terms mean.  (Does "total output power"
include far IR and heat?)

maybe it has some real bright LEDs to indicate it's on?


Message: 3
Date: Tue, 2 Oct 2012 22:33:28 -0400 (EDT)
From: johncroos@aol.com
To: time-nuts@febo.com
Subject: Re: [time-nuts] GPS Jammer
Message-ID: 8CF6F269AD88F67-1D20-7F60D@webmail-d139.sysops.aol.com
Content-Type: text/plain; charset="us-ascii"; format=flowed

In considering the effect of a simple jammer on a GPS receiver, a
simple link analysis
is insufficient.

What must also be considered is the anti-jam capability of the receiver
which due to spread spectrum processing gain will reject any simple
jamming signal even though is it 10's of dB stronger than the desired
signal.

73 -john k6iql


Message: 4
Date: Tue, 02 Oct 2012 19:34:47 -0700
From: Jim Lux jimlux@earthlink.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] Best counter setting for ADEV?
Message-ID: 506BA447.5090704@earthlink.net
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

On 10/2/12 7:00 PM, shalimr9@gmail.com wrote:

The Thunderbolt is a special case that does not provide sawtooth

correction
because it does not need it.

It uses the OCXO as the clock for the processor while disciplining it

to GPS
so there is no nominal timing error between where the 1PPS is versus
where it
should be.

The processor is able to bring the PPS edge exactly where it wants

it, instead
of the typical 25 to 40 ns granularity of most other GPS receivers that
operate
on a separate clock.

Pretty simple and elegant solution.

But it does depend on having a good oscillator that can be shoved
around.  That costs money and power.


Message: 5
Date: Tue, 02 Oct 2012 19:37:32 -0700
From: Jim Lux jimlux@earthlink.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] RFX GPSDO - Anybody played with one of these?
Message-ID: 506BA4EC.2070403@earthlink.net
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

On 10/2/12 2:36 PM, SAIDJACK@aol.com wrote:

Hello Paul,

thanks much for the feedback!

Yes, we think we have identified a nice combination of oscillators,

GPS,

and firmware that seems to work pretty well. The GPSTCXO units cannot

be

compared to a lower cost $150 Thunderbolt in terms of phase noise or

stability

of course, and they have CMOS 10MHz outputs, but then the  TB's cost

around

$1500 new I guess.

We think the GPSTCXO's and LC_XO type units will work quite well

wherever

standard OCXO's are used today, and power/size/weight are an  issue.

Intriguing.. Can it handle the Doppler, etc., for a cubesat in LEO?
(7km/s)  The total Doppler isn't usually the issue (the GPS satellites
are moving faster, after all), but the receiver may not work for high
velocities, high altitudes?


Message: 6
Date: Tue, 2 Oct 2012 21:37:53 -0500
From: Bill Dailey docdailey@gmail.com
To: Time Nuts time-nuts@febo.com
Subject: [time-nuts] Centering ocxo
Message-ID: 21F2B49D-72FA-46A5-956E-B45ACBD6EEEB@gmail.com
Content-Type: text/plain; charset=us-ascii

I am ok for awhile but how do you center the efc of an ocxo?  I
understand there
is something (screw) to adjust the ocxo so it is approximately on freq
with 2.5v
efc.

Specific oscillator datum-1111c. I have he datasheet but doesn't say
"coarse
frequency adjust this screw" or some such.

Doc
KX0O

Sent from my iPad


Message: 7
Date: Tue, 02 Oct 2012 19:45:32 -0700
From: Jim Lux jimlux@earthlink.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] GPS Jammer
Message-ID: 506BA6CC.2040702@earthlink.net
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

On 10/2/12 7:33 PM, johncroos@aol.com wrote:

In considering the effect of a simple jammer on a GPS receiver, a

simple

link analysis
is insufficient.

What must also be considered is the anti-jam capability of the

receiver

which due to spread spectrum processing gain will reject any simple
jamming signal even though is it 10's of dB stronger than the desired
signal.

not most simple GPS receivers which have very little AJ capability.
They
have a single bit quantizer (or maybe a 1.5 or 2 bit) after the LNA.
If
the LNA doesn't saturate, then the quantizer is captured by the strong
CW carrier.

This is a classic problem with DSSS receivers and led to a lot of
research in the 80s on things like "adaptive excisers" to remove CW
carriers.

If you built a linear receiver with a lot of dynamic range, then, yes,
the process gain will suppress the CW tone, but you still have to
acquire the code, and as Dixon says (paraphrasing) "acquisition is the
secret sauce in spread spectrum systems".  Back when I was doing this
kind of thing seriously (mid to late 80s), acquisition, particularly
robust techniques, were literally SECRET (in the DoD sense).

There have been a nice series of articles in GPS World over the past
few
months about the variety of inexpensive GPS jammers out there. (and the
problems they cause).


Message: 8
Date: Tue, 2 Oct 2012 22:53:18 -0400
From: paul swed paulswedb@gmail.com
To: jfor@quikus.com, Discussion of precise time and frequency
measurement time-nuts@febo.com
Subject: Re: [time-nuts] Google's Spanner uses GPS and atomic time
Message-ID:
CAD2JfAi3Q_BgbBhSGK3dfz_kdG2-_3vmhAkNv0pQJbaaS19KNA@mail.gmail.com
Content-Type: text/plain; charset=ISO-8859-1

And yes thank god we are cutting funding to those pesky satellites.
Seems
we can't afford them anymore. But fortunately other countries are
filling
our gap slow but surely.
There was an article this month in GNSS about the funding cuts.
Regards
Paul
WB8TSL

On Tue, Oct 2, 2012 at 9:11 PM, J. Forster jfor@quikus.com wrote:

The US, foolishly IMO, has killed LORAN-C and is killing WWVB. Hence

my

comment. Essentially, GPS is soon going to be the sole source of a
standard of time interval.

It is going to take a disaster of some kind to return sanity.

-John

===========

John writes:

Seems like a good reason to have LORAN-C or some other

backup/sanity

check.

What LORAN? I thought the U.S. had shut down all LORAN transmissions
in order to enhance the vulnerability of navigation systems in the
U.S. (?).

--
Anthony


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

End of time-nuts Digest, Vol 99, Issue 19


Hi Jim - Thanks for the update on the modern GPS receivers. I was aware that the modern ones do not have a classical analog tracking loop, much less a bunch of them. However it is a useful concept for purposes of explanation that you do not need the 1 pps to lock up the 10 MHz VCXO - which was my main point. The Tbolt block diagram in the manual Figure 5-10 shows the 10 MHz VCXO output going to the receiver and also to the output. The 1 pps comes from the "cpu and support" circuit. While the diagram is clearly simplistic - it implies that the Receiver circuitry first locks up the 10 MHz oscillator and the the 1 pps is derived from that. My only point was that it is possible and perhaps even better to discipline an oscillator using a code correlator (however implemented) rather than steering using the 1 pps. I believe this is why the T bolts work so well. For one thing the loop should work better and faster if the input is at the chipping rate rather than at 1 pps as the information rate is higher. On doppler - I believe that since the spread spectrum sidebands are coherent with the suppressed carrier, their relationship to it is unchanged by doppler and thus it should be possible to achieve a code correlation on a doppler shifted signal. The recovered carrier would be shifted in frequency by the doppler but it would still be recovered - at least by a classical IF correlator. - On jamming - maybe so, but the effect of the receiving correlator is to spread the energy of a CW interferer and concentrate the energy of the signal with the matching PN modulation is it not???? good fun anyway! -73 john k6iql -----Original Message----- From: time-nuts-request <time-nuts-request@febo.com> To: time-nuts <time-nuts@febo.com> Sent: Tue, Oct 2, 2012 9:53 pm Subject: time-nuts Digest, Vol 99, Issue 19 Send time-nuts mailing list submissions to time-nuts@febo.com To subscribe or unsubscribe via the World Wide Web, visit https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts or, via email, send a message with subject or body 'help' to time-nuts-request@febo.com You can reach the person managing the list at time-nuts-owner@febo.com When replying, please edit your Subject line so it is more specific than "Re: Contents of time-nuts digest..." Today's Topics: 1. Re: GPS Modulation and 10 MHz Delay Lock (Jim Lux) 2. Re: GPS Jammer (Jim Lux) 3. Re: GPS Jammer (johncroos@aol.com) 4. Re: Best counter setting for ADEV? (Jim Lux) 5. Re: RFX GPSDO - Anybody played with one of these? (Jim Lux) 6. Centering ocxo (Bill Dailey) 7. Re: GPS Jammer (Jim Lux) 8. Re: Google's Spanner uses GPS and atomic time (paul swed) ---------------------------------------------------------------------- Message: 1 Date: Tue, 02 Oct 2012 19:30:20 -0700 From: Jim Lux <jimlux@earthlink.net> To: time-nuts@febo.com Subject: Re: [time-nuts] GPS Modulation and 10 MHz Delay Lock Message-ID: <506BA33C.1010606@earthlink.net> Content-Type: text/plain; charset=ISO-8859-1; format=flowed On 10/2/12 3:39 PM, johncroos@aol.com wrote: > Hello All - > > Here is a link that describes the GPS modulation. You do not need the > 1 pps to lock the 10 MHz oscillator to the atomic clock in the satellites. > > http://www.kowoma.de/en/gps/signals.htm > > If you look at the block diagram you see PN code modulates the carrier at > the 1.023 MHz chip rate. This is done by BPSK modulation of the carrier > with the PN code. It can be done simply with a double balanced mixer. > > This spreads the signal with PSK at the chip( i.e. code clock) rate. > > Note also the modulo - 2 addition of the data to the code sequence. This > called code inversion > modulation. After de-spread of the code in the receiver - the signal is > then simple BPSK and > may be demodulated by a Costas or Squaring Loop to get at the data message. > > The obtain precision frequency needed I believe the T bolt simply locks > to the chipping rate > using some form of Delay Lock Loop. It is NOT at PLL. There is no need > what ever to > deal with the 1 pps using this method. The internal 10 MHz oscillator is > controlled by this locking circuit and > is part of the code correlation loop. That's not quite how it works.. It would work for terrestrial links where there is no Doppler, but in the GPS case, there is significant Doppler shift on all the signals. Since the carrier and the chips are generated from a common source on the spacecraft (the carrier frequency is a multiple of the chip rate, in fact), you can recover carrier and chips at the same time. But.. most receivers these days don't actually have an analog tracking loop at all. They digitize the input signal (1 bit quantizer) at a rate that makes the carrier alias down to something convenient (a few hundred kHz is typical.. you want it far enough away from zero that Doppler never makes it go negative). In the experimental receiver in SCaN Testbed flying on ISS it's about 39 MHz sample rate. Once you've got your one bit samples, you do some sort of combined Doppler/Code phase acquisition (these days, often using an FFT), then track both together digitally using some form of NCO. The tracking loops for all the satellite signals aren't necessarily independent and might be part of a Kalman filter that estimates all the observables together. Finally, from all that, you have an estimate of your local clock offset and timing offset, and from that you can generate your 1pps, typically with another NCO (with granularity of your clock rate). Since it's unlikely that your clock is EXACTLY an even number of cycles per second, at each second, a bit of error accumulates, until you have an whole cycle's worth leading to the familiar sawtooth error. That sawtooth error is predictable, of course, so you can generate a "time error" estimate for each 1pps pulse (or, even, control a variable delay to line it up). The important thing is that in modern receivers, nowhere is there a signal at the GPS carrier frequency, nor is there a signal at the chip rate. There *is* probably a signal (with low precision) at the code epoch (every millisecond), but it's different for each satellite signal, of course. ------------------------------ Message: 2 Date: Tue, 02 Oct 2012 19:32:48 -0700 From: Jim Lux <jimlux@earthlink.net> To: time-nuts@febo.com Subject: Re: [time-nuts] GPS Jammer Message-ID: <506BA3D0.4090609@earthlink.net> Content-Type: text/plain; charset=ISO-8859-1; format=flowed On 10/2/12 4:48 PM, Charles P. Steinmetz wrote: > >> The Power Output is 0.5 Watts and it claims a jamming range of 1-10 >> Meters. >> Anybody think there is something wrong? > > I'd expect a much greater range with a 0.5 W jammer. But note that 0.5 > W is the "total output power" -- the "transmit power" is only 10 dBm > (0.01 W). Whatever those terms mean. (Does "total output power" > include far IR and heat?) > maybe it has some real bright LEDs to indicate it's on? ------------------------------ Message: 3 Date: Tue, 2 Oct 2012 22:33:28 -0400 (EDT) From: johncroos@aol.com To: time-nuts@febo.com Subject: Re: [time-nuts] GPS Jammer Message-ID: <8CF6F269AD88F67-1D20-7F60D@webmail-d139.sysops.aol.com> Content-Type: text/plain; charset="us-ascii"; format=flowed In considering the effect of a simple jammer on a GPS receiver, a simple link analysis is insufficient. What must also be considered is the anti-jam capability of the receiver which due to spread spectrum processing gain will reject any simple jamming signal even though is it 10's of dB stronger than the desired signal. 73 -john k6iql ------------------------------ Message: 4 Date: Tue, 02 Oct 2012 19:34:47 -0700 From: Jim Lux <jimlux@earthlink.net> To: time-nuts@febo.com Subject: Re: [time-nuts] Best counter setting for ADEV? Message-ID: <506BA447.5090704@earthlink.net> Content-Type: text/plain; charset=ISO-8859-1; format=flowed On 10/2/12 7:00 PM, shalimr9@gmail.com wrote: > The Thunderbolt is a special case that does not provide sawtooth correction because it does not need it. > > It uses the OCXO as the clock for the processor while disciplining it to GPS so there is no nominal timing error between where the 1PPS is versus where it should be. > > The processor is able to bring the PPS edge exactly where it wants it, instead of the typical 25 to 40 ns granularity of most other GPS receivers that operate on a separate clock. > > Pretty simple and elegant solution. > But it does depend on having a good oscillator that can be shoved around. That costs money and power. ------------------------------ Message: 5 Date: Tue, 02 Oct 2012 19:37:32 -0700 From: Jim Lux <jimlux@earthlink.net> To: time-nuts@febo.com Subject: Re: [time-nuts] RFX GPSDO - Anybody played with one of these? Message-ID: <506BA4EC.2070403@earthlink.net> Content-Type: text/plain; charset=ISO-8859-1; format=flowed On 10/2/12 2:36 PM, SAIDJACK@aol.com wrote: > Hello Paul, > > thanks much for the feedback! > > Yes, we think we have identified a nice combination of oscillators, GPS, > and firmware that seems to work pretty well. The GPSTCXO units cannot be > compared to a lower cost $150 Thunderbolt in terms of phase noise or stability > of course, and they have CMOS 10MHz outputs, but then the TB's cost around > $1500 new I guess. > > We think the GPSTCXO's and LC_XO type units will work quite well wherever > standard OCXO's are used today, and power/size/weight are an issue. > Intriguing.. Can it handle the Doppler, etc., for a cubesat in LEO? (7km/s) The total Doppler isn't usually the issue (the GPS satellites are moving faster, after all), but the receiver may not work for high velocities, high altitudes? ------------------------------ Message: 6 Date: Tue, 2 Oct 2012 21:37:53 -0500 From: Bill Dailey <docdailey@gmail.com> To: Time Nuts <time-nuts@febo.com> Subject: [time-nuts] Centering ocxo Message-ID: <21F2B49D-72FA-46A5-956E-B45ACBD6EEEB@gmail.com> Content-Type: text/plain; charset=us-ascii I am ok for awhile but how do you center the efc of an ocxo? I understand there is something (screw) to adjust the ocxo so it is approximately on freq with 2.5v efc. Specific oscillator datum-1111c. I have he datasheet but doesn't say "coarse frequency adjust this screw" or some such. Doc KX0O Sent from my iPad ------------------------------ Message: 7 Date: Tue, 02 Oct 2012 19:45:32 -0700 From: Jim Lux <jimlux@earthlink.net> To: time-nuts@febo.com Subject: Re: [time-nuts] GPS Jammer Message-ID: <506BA6CC.2040702@earthlink.net> Content-Type: text/plain; charset=ISO-8859-1; format=flowed On 10/2/12 7:33 PM, johncroos@aol.com wrote: > In considering the effect of a simple jammer on a GPS receiver, a simple > link analysis > is insufficient. > > What must also be considered is the anti-jam capability of the receiver > which due to spread spectrum processing gain will reject any simple > jamming signal even though is it 10's of dB stronger than the desired > signal. not most simple GPS receivers which have very little AJ capability. They have a single bit quantizer (or maybe a 1.5 or 2 bit) after the LNA. If the LNA doesn't saturate, then the quantizer is captured by the strong CW carrier. This is a classic problem with DSSS receivers and led to a lot of research in the 80s on things like "adaptive excisers" to remove CW carriers. If you built a linear receiver with a lot of dynamic range, then, yes, the process gain will suppress the CW tone, but you still have to acquire the code, and as Dixon says (paraphrasing) "acquisition is the secret sauce in spread spectrum systems". Back when I was doing this kind of thing seriously (mid to late 80s), acquisition, particularly robust techniques, were literally SECRET (in the DoD sense). There have been a nice series of articles in GPS World over the past few months about the variety of inexpensive GPS jammers out there. (and the problems they cause). ------------------------------ Message: 8 Date: Tue, 2 Oct 2012 22:53:18 -0400 From: paul swed <paulswedb@gmail.com> To: jfor@quikus.com, Discussion of precise time and frequency measurement <time-nuts@febo.com> Subject: Re: [time-nuts] Google's Spanner uses GPS and atomic time Message-ID: <CAD2JfAi3Q_BgbBhSGK3dfz_kdG2-_3vmhAkNv0pQJbaaS19KNA@mail.gmail.com> Content-Type: text/plain; charset=ISO-8859-1 And yes thank god we are cutting funding to those pesky satellites. Seems we can't afford them anymore. But fortunately other countries are filling our gap slow but surely. There was an article this month in GNSS about the funding cuts. Regards Paul WB8TSL On Tue, Oct 2, 2012 at 9:11 PM, J. Forster <jfor@quikus.com> wrote: > The US, foolishly IMO, has killed LORAN-C and is killing WWVB. Hence my > comment. Essentially, GPS is soon going to be the sole source of a > standard of time interval. > > It is going to take a disaster of some kind to return sanity. > > -John > > =========== > > > > > John writes: > > > >> Seems like a good reason to have LORAN-C or some other backup/sanity > >> check. > > > > What LORAN? I thought the U.S. had shut down all LORAN transmissions > > in order to enhance the vulnerability of navigation systems in the > > U.S. (?). > > > > -- > > Anthony > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > ------------------------------ _______________________________________________ time-nuts mailing list time-nuts@febo.com https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts End of time-nuts Digest, Vol 99, Issue 19 *****************************************
JL
Jim Lux
Wed, Oct 3, 2012 4:26 AM

On 10/2/12 8:40 PM, johncroos@aol.com wrote:

Hi Jim -

Thanks for the update on the modern GPS receivers. I was aware that the
modern ones
do not have a classical analog tracking loop, much less a bunch of them.
However it is a useful concept for purposes of explanation that you do
not need the 1 pps to
lock up the 10 MHz VCXO - which was my main point.

The Tbolt block diagram in the manual Figure 5-10 shows the 10 MHz VCXO
output going to
the receiver and also to the output. The 1 pps comes from the "cpu and
support" circuit.

yes.. the Tbolt is an example of a receiver that shifts the frequency of
the reference so that it can generate an "on the mark" 1pps (and
generate an accurate 10 MHz as well...)

  • On jamming - maybe so, but the effect of the receiving correlator is
    to spread the energy of a CW interferer
    and concentrate the energy of the signal with the matching PN modulation
    is it not????

Yes, if the receiver is linear (e.g. say you do a sliding code
correlator and slide until you get the peak, with the correlator using a
multiplier)...

But since most (inexpensive) receivers have hard limiters/1 bit
quantizers in front of the correlator, what goes into the correlator is
a square wave at the jammer frequency, and you can slide your PN code
all day and not get a peak.

There's some analysis out there that tells you how many bits you need
for a given Jammer/Signal ratio, but in general, if the jammer is 20 dB
over the signal, you need 3-4 bits.

Back in the day, when bits were very expensive, that's why narrow band
excisers were popular.. basically you'd run something like a PLL to
recover the tone jammer, and subtract it out.

On 10/2/12 8:40 PM, johncroos@aol.com wrote: > Hi Jim - > > Thanks for the update on the modern GPS receivers. I was aware that the > modern ones > do not have a classical analog tracking loop, much less a bunch of them. > However it is a useful concept for purposes of explanation that you do > not need the 1 pps to > lock up the 10 MHz VCXO - which was my main point. > > The Tbolt block diagram in the manual Figure 5-10 shows the 10 MHz VCXO > output going to > the receiver and also to the output. The 1 pps comes from the "cpu and > support" circuit. yes.. the Tbolt is an example of a receiver that shifts the frequency of the reference so that it can generate an "on the mark" 1pps (and generate an accurate 10 MHz as well...) > > - On jamming - maybe so, but the effect of the receiving correlator is > to spread the energy of a CW interferer > and concentrate the energy of the signal with the matching PN modulation > is it not???? Yes, if the receiver is linear (e.g. say you do a sliding code correlator and slide until you get the peak, with the correlator using a multiplier)... But since most (inexpensive) receivers have hard limiters/1 bit quantizers in front of the correlator, what goes into the correlator is a square wave at the jammer frequency, and you can slide your PN code all day and not get a peak. There's some analysis out there that tells you how many bits you need for a given Jammer/Signal ratio, but in general, if the jammer is 20 dB over the signal, you need 3-4 bits. Back in the day, when bits were very expensive, that's why narrow band excisers were popular.. basically you'd run something like a PLL to recover the tone jammer, and subtract it out.