volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Seek full Manual - EDC DC millivolt calibrator E100C

C
cfo
Sun, Mar 17, 2013 7:32 PM

Guyzz

I just bought a EDC DC millivolt calibrator E100C ,
is $215  a fair price ?

I expect to  get it  this week.

I think it's the same as the Kron-Hite E100RC.

The KronHite-Manual says it stops after section3
(and $$ for the full manual)  , the other has to section 4.

See http://tinyurl.com/capynjd

So i was wondering anyone has the full manual for either the KH E100R
or even better the EDC E100C

Any hints or quirks , also much appreciated.

I have an Agilent 34401A , and a Geller 10v ref (6 month old) , so i hope
i can verify it.

CFO
Denmark

Guyzz I just bought a EDC DC millivolt calibrator E100C , is $215 a fair price ? I expect to get it this week. I think it's the same as the Kron-Hite E100RC. The KronHite-Manual says it stops after section3 (and $$ for the full manual) , the other has to section 4. See http://tinyurl.com/capynjd So i was wondering anyone has the full manual for either the KH E100R or even better the EDC E100C Any hints or quirks , also much appreciated. I have an Agilent 34401A , and a Geller 10v ref (6 month old) , so i hope i can verify it. CFO Denmark
JL
J. L. Trantham
Mon, Mar 18, 2013 1:16 AM

CFO,

Several months ago, I bought an EDC CR103 DC Current and Voltage Standard
and ran into the same problem.

I emailed Kron-Hite, asking if the manual was available, if so, how much,
etc.  The reply was an emailed .PDF complete manual, schematics and
everything.

You might try asking them if they still have the manual available, etc.
Couldn't hurt.

Good luck.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of cfo
Sent: Sunday, March 17, 2013 2:33 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C

Guyzz

I just bought a EDC DC millivolt calibrator E100C ,
is $215  a fair price ?

I expect to  get it  this week.

I think it's the same as the Kron-Hite E100RC.

The KronHite-Manual says it stops after section3
(and $$ for the full manual)  , the other has to section 4.

See http://tinyurl.com/capynjd

So i was wondering anyone has the full manual for either the KH E100R
or even better the EDC E100C

Any hints or quirks , also much appreciated.

I have an Agilent 34401A , and a Geller 10v ref (6 month old) , so i hope
i can verify it.

CFO
Denmark


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

CFO, Several months ago, I bought an EDC CR103 DC Current and Voltage Standard and ran into the same problem. I emailed Kron-Hite, asking if the manual was available, if so, how much, etc. The reply was an emailed .PDF complete manual, schematics and everything. You might try asking them if they still have the manual available, etc. Couldn't hurt. Good luck. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of cfo Sent: Sunday, March 17, 2013 2:33 PM To: volt-nuts@febo.com Subject: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C Guyzz I just bought a EDC DC millivolt calibrator E100C , is $215 a fair price ? I expect to get it this week. I think it's the same as the Kron-Hite E100RC. The KronHite-Manual says it stops after section3 (and $$ for the full manual) , the other has to section 4. See http://tinyurl.com/capynjd So i was wondering anyone has the full manual for either the KH E100R or even better the EDC E100C Any hints or quirks , also much appreciated. I have an Agilent 34401A , and a Geller 10v ref (6 month old) , so i hope i can verify it. CFO Denmark _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
CB
Charles Black
Mon, Mar 18, 2013 2:29 AM

Hi CFO,

Too bad your so far away since I wouldn't mind comparing your 10v
standard against my Datron 4910 just for the fun of it.. The 4910 was
calibrated by Fluke about a year ago. All four channels ha.ve been
stable within a few microvolts in my lab over the last year. So far it
hasn't been compared to the outside world though and I really  should do
some comparisons to be sure that everything is OK.

Charlie

On 3/17/2013 12:32 PM, cfo wrote:

Guyzz

I just bought a EDC DC millivolt calibrator E100C ,
is $215  a fair price ?

I expect to  get it  this week.

I think it's the same as the Kron-Hite E100RC.

The KronHite-Manual says it stops after section3
(and $$ for the full manual)  , the other has to section 4.

See http://tinyurl.com/capynjd

So i was wondering anyone has the full manual for either the KH E100R
or even better the EDC E100C

Any hints or quirks , also much appreciated.

I have an Agilent 34401A , and a Geller 10v ref (6 month old) , so i hope
i can verify it.

CFO
Denmark


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi CFO, Too bad your so far away since I wouldn't mind comparing your 10v standard against my Datron 4910 just for the fun of it.. The 4910 was calibrated by Fluke about a year ago. All four channels ha.ve been stable within a few microvolts in my lab over the last year. So far it hasn't been compared to the outside world though and I really should do some comparisons to be sure that everything is OK. Charlie On 3/17/2013 12:32 PM, cfo wrote: > Guyzz > > I just bought a EDC DC millivolt calibrator E100C , > is $215 a fair price ? > > I expect to get it this week. > > I think it's the same as the Kron-Hite E100RC. > > The KronHite-Manual says it stops after section3 > (and $$ for the full manual) , the other has to section 4. > > See http://tinyurl.com/capynjd > > So i was wondering anyone has the full manual for either the KH E100R > or even better the EDC E100C > > Any hints or quirks , also much appreciated. > > I have an Agilent 34401A , and a Geller 10v ref (6 month old) , so i hope > i can verify it. > > CFO > Denmark > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > >
C
cfo
Mon, Mar 18, 2013 8:21 PM

On Sun, 17 Mar 2013 20:16:29 -0500, J. L. Trantham wrote:

CFO,

Several months ago, I bought an EDC CR103 DC Current and Voltage
Standard and ran into the same problem.

I emailed Kron-Hite, asking if the manual was available, if so, how
much,
etc.  The reply was an emailed .PDF complete manual, schematics and
everything.

You might try asking them if they still have the manual available, etc.
Couldn't hurt.

Good luck.

Joe

Hi Joe

Thank you for the tip.
I followed your advice , and emailed Kron-Hite service , asking if it
would be possible to get a copy of the service manual.

And i just got a very plesant response from
their Service Department Manager , with the complete E100C service manual
attached. and a best of luck with the restoration project.

Thumbs up Krohn-Hite , and thank you for the excellent service.

CFO - Denmark

On Sun, 17 Mar 2013 20:16:29 -0500, J. L. Trantham wrote: > CFO, > > Several months ago, I bought an EDC CR103 DC Current and Voltage > Standard and ran into the same problem. > > I emailed Kron-Hite, asking if the manual was available, if so, how > much, > etc. The reply was an emailed .PDF complete manual, schematics and > everything. > > You might try asking them if they still have the manual available, etc. > Couldn't hurt. > > Good luck. > > Joe > Hi Joe Thank you for the tip. I followed your advice , and emailed Kron-Hite service , asking if it would be possible to get a copy of the service manual. And i just got a very plesant response from their Service Department Manager , with the complete E100C service manual attached. and a best of luck with the restoration project. Thumbs up Krohn-Hite , and thank you for the excellent service. CFO - Denmark
JL
John Lofgren
Mon, Mar 18, 2013 8:27 PM

I had a similar experience with an old K-H function generator.  I just called them up and they e-mailed me a scan of the manual.  A thumbs-up from me, too.

Other test equipment vendors should take note.

-John

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of cfo
Sent: Monday, March 18, 2013 3:21 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C

On Sun, 17 Mar 2013 20:16:29 -0500, J. L. Trantham wrote:

CFO,

Several months ago, I bought an EDC CR103 DC Current and Voltage
Standard and ran into the same problem.

I emailed Kron-Hite, asking if the manual was available, if so, how
much,
etc.  The reply was an emailed .PDF complete manual, schematics and
everything.

You might try asking them if they still have the manual available, etc.
Couldn't hurt.

Good luck.

Joe

Hi Joe

Thank you for the tip.
I followed your advice , and emailed Kron-Hite service , asking if it
would be possible to get a copy of the service manual.

And i just got a very plesant response from
their Service Department Manager , with the complete E100C service manual
attached. and a best of luck with the restoration project.

Thumbs up Krohn-Hite , and thank you for the excellent service.

CFO - Denmark


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I had a similar experience with an old K-H function generator. I just called them up and they e-mailed me a scan of the manual. A thumbs-up from me, too. Other test equipment vendors should take note. -John -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of cfo Sent: Monday, March 18, 2013 3:21 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C On Sun, 17 Mar 2013 20:16:29 -0500, J. L. Trantham wrote: > CFO, > > Several months ago, I bought an EDC CR103 DC Current and Voltage > Standard and ran into the same problem. > > I emailed Kron-Hite, asking if the manual was available, if so, how > much, > etc. The reply was an emailed .PDF complete manual, schematics and > everything. > > You might try asking them if they still have the manual available, etc. > Couldn't hurt. > > Good luck. > > Joe > Hi Joe Thank you for the tip. I followed your advice , and emailed Kron-Hite service , asking if it would be possible to get a copy of the service manual. And i just got a very plesant response from their Service Department Manager , with the complete E100C service manual attached. and a best of luck with the restoration project. Thumbs up Krohn-Hite , and thank you for the excellent service. CFO - Denmark _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JL
J. L. Trantham
Tue, Mar 19, 2013 12:27 AM

CFO,

Glad it worked.  You might want to post it on some site or the group list or
both, in order to help others that will follow.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of cfo
Sent: Monday, March 18, 2013 3:21 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator
E100C

On Sun, 17 Mar 2013 20:16:29 -0500, J. L. Trantham wrote:

CFO,

Several months ago, I bought an EDC CR103 DC Current and Voltage
Standard and ran into the same problem.

I emailed Kron-Hite, asking if the manual was available, if so, how
much, etc.  The reply was an emailed .PDF complete manual, schematics
and everything.

You might try asking them if they still have the manual available, etc.
Couldn't hurt.

Good luck.

Joe

Hi Joe

Thank you for the tip.
I followed your advice , and emailed Kron-Hite service , asking if it would
be possible to get a copy of the service manual.

And i just got a very plesant response from their Service Department Manager
, with the complete E100C service manual attached. and a best of luck with
the restoration project.

Thumbs up Krohn-Hite , and thank you for the excellent service.

CFO - Denmark


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

CFO, Glad it worked. You might want to post it on some site or the group list or both, in order to help others that will follow. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of cfo Sent: Monday, March 18, 2013 3:21 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C On Sun, 17 Mar 2013 20:16:29 -0500, J. L. Trantham wrote: > CFO, > > Several months ago, I bought an EDC CR103 DC Current and Voltage > Standard and ran into the same problem. > > I emailed Kron-Hite, asking if the manual was available, if so, how > much, etc. The reply was an emailed .PDF complete manual, schematics > and everything. > > You might try asking them if they still have the manual available, etc. > Couldn't hurt. > > Good luck. > > Joe > Hi Joe Thank you for the tip. I followed your advice , and emailed Kron-Hite service , asking if it would be possible to get a copy of the service manual. And i just got a very plesant response from their Service Department Manager , with the complete E100C service manual attached. and a best of luck with the restoration project. Thumbs up Krohn-Hite , and thank you for the excellent service. CFO - Denmark _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
C
cfo
Tue, Mar 19, 2013 12:51 PM

On Mon, 18 Mar 2013 19:27:28 -0500, J. L. Trantham wrote:

CFO,

Glad it worked.  You might want to post it on some site or the group
list or both, in order to help others that will follow.

Joe

I thought of that , but i wonder what KH says if i make the manual public.
I wouldn't like to ruin the nice service for other volt-nuts.

If there is any wishes from the group , i'll ask KH if it would be ok to
upload it to ie. didiers site : http://www.ko4bb.com/manuals.php

But i think i owe to ask first.

CFO

On Mon, 18 Mar 2013 19:27:28 -0500, J. L. Trantham wrote: > CFO, > > Glad it worked. You might want to post it on some site or the group > list or both, in order to help others that will follow. > > Joe I thought of that , but i wonder what KH says if i make the manual public. I wouldn't like to ruin the nice service for other volt-nuts. If there is any wishes from the group , i'll ask KH if it would be ok to upload it to ie. didiers site : http://www.ko4bb.com/manuals.php But i think i owe to ask first. CFO
JL
J. L. Trantham
Tue, Mar 19, 2013 1:05 PM

Agreed.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of cfo
Sent: Tuesday, March 19, 2013 7:52 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator
E100C

On Mon, 18 Mar 2013 19:27:28 -0500, J. L. Trantham wrote:

CFO,

Glad it worked.  You might want to post it on some site or the group
list or both, in order to help others that will follow.

Joe

I thought of that , but i wonder what KH says if i make the manual public.
I wouldn't like to ruin the nice service for other volt-nuts.

If there is any wishes from the group , i'll ask KH if it would be ok to
upload it to ie. didiers site : http://www.ko4bb.com/manuals.php

But i think i owe to ask first.

CFO


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Agreed. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of cfo Sent: Tuesday, March 19, 2013 7:52 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C On Mon, 18 Mar 2013 19:27:28 -0500, J. L. Trantham wrote: > CFO, > > Glad it worked. You might want to post it on some site or the group > list or both, in order to help others that will follow. > > Joe I thought of that , but i wonder what KH says if i make the manual public. I wouldn't like to ruin the nice service for other volt-nuts. If there is any wishes from the group , i'll ask KH if it would be ok to upload it to ie. didiers site : http://www.ko4bb.com/manuals.php But i think i owe to ask first. CFO _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
C
cfo
Tue, Mar 19, 2013 1:45 PM

On Tue, 19 Mar 2013 08:05:54 -0500, J. L. Trantham wrote:

Agreed.

Joe

...
...
...

If there is any wishes from the group , i'll ask KH if it would be ok to
upload it to ie. didiers site : http://www.ko4bb.com/manuals.php

But i think i owe to ask first.

CFO

I have asked KH , and they prefer that it would not be uploaded to the
"Net".

I will respect that decision , and it doesn't seem to be a problem to get
it of one asks nicely in an email.

CFO

On Tue, 19 Mar 2013 08:05:54 -0500, J. L. Trantham wrote: > Agreed. > > Joe ... ... ... > If there is any wishes from the group , i'll ask KH if it would be ok to > upload it to ie. didiers site : http://www.ko4bb.com/manuals.php > > But i think i owe to ask first. > > CFO I have asked KH , and they prefer that it would not be uploaded to the "Net". I will respect that decision , and it doesn't seem to be a problem to get it of one asks nicely in an email. CFO
JL
J. L. Trantham
Tue, Mar 19, 2013 1:54 PM

If we make note of this on the reflector, others will be able to follow the
same path or ask us about how it was done.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of cfo
Sent: Tuesday, March 19, 2013 8:46 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator
E100C

On Tue, 19 Mar 2013 08:05:54 -0500, J. L. Trantham wrote:

Agreed.

Joe

...
...
...

If there is any wishes from the group , i'll ask KH if it would be ok
to upload it to ie. didiers site : http://www.ko4bb.com/manuals.php

But i think i owe to ask first.

CFO

I have asked KH , and they prefer that it would not be uploaded to the
"Net".

I will respect that decision , and it doesn't seem to be a problem to get
it of one asks nicely in an email.

CFO


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

If we make note of this on the reflector, others will be able to follow the same path or ask us about how it was done. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of cfo Sent: Tuesday, March 19, 2013 8:46 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Seek full Manual - EDC DC millivolt calibrator E100C On Tue, 19 Mar 2013 08:05:54 -0500, J. L. Trantham wrote: > Agreed. > > Joe ... ... ... > If there is any wishes from the group , i'll ask KH if it would be ok > to upload it to ie. didiers site : http://www.ko4bb.com/manuals.php > > But i think i owe to ask first. > > CFO I have asked KH , and they prefer that it would not be uploaded to the "Net". I will respect that decision , and it doesn't seem to be a problem to get it of one asks nicely in an email. CFO _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.