Hi Gerd and Charles,
Thanks for the input it has given me other things to think about.
I checked the output of pin 6 of U1 and I get a voltage variation of 10.2 mv
from +15 to -15v adjustment on R6, so that appears to be working correctly.
Now the fun starts :- using my HP 3478A to measure the dc on the output
terminals I get  a reading of 0.9mv on the 1 volt range, 0.3 mv on the 10
volt range, and 1.8 mv on the 100volt range.
As a quick check of wether C8 may be leaky I pulled Q38 out of it's socket
and got a constant DC out on the output terminals of 0.22mv on all three
ranges, so maybe C8 is leaky?
Using a fluke 77 3.5 digit multimeter yields vastly different DC readings,
up to 5.5 v on the 100 volt range, so I suspect the Fluke 77 is reading some
of the AC component on the output terminals even though DC is selected.
Last night I did check the input and output on the amplifier, and on the 1 v
range I was getting an input of 1v and an output of 3 volts, so the gain was
correct, I did not check it on the 100v range though to see if the gain was
30.
The input and output was a good clean sine wave on the amplifier.
I have previously checked the 190 volt + and - supplies and they were spot
on, but I will check them again before I pull the amplifier board apart
again to remove C1 and C8 to test with a higher voltage other than the
battery voltage of the multimeter, to see if they may be leaking at a higher
voltage.
Thanks,
Ken


Message: 2
Date: Wed, 09 Jul 2014 14:21:13 -0400
From: Charles Steinmetz csteinmetz@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 5200A repair
Message-ID: 20140709222123.LMBGV5Je@smtp2o.mail.yandex.net
Content-Type: text/plain; charset="us-ascii"; format=flowed

Ken wrote:

R6 slider to U1 varies from +15v to -15v but the output of U1 only
varies by mv's, I put this down to the loading of R9 to earth on the
input to pin3 of U1 (221 ohm),only allowing a small variation to the
input on pin
3 of U1, maybe this needs further investigation!

Look at the circuit.  The 100k potentiometer (R6) feeds a voltage divider
composed of R8 and R9 (464k and 221 ohms, respectively).  The tap of this
divider feeds the reference input of U1.  R8 and R9 attenuate the +/- 15v
range of R6 by a factor of 210, so the divider tap voltage (i.e., the
reference voltage for U1) only varies +/- 7.14 mV from ground for the full
range of R6.

U1 is an integrator, so it has lots of gain at DC (Q38 switches the
integrating capacitor to give two time constants).  If one of these
capacitors is leaky or shorted (or there is another leakage path from Pin 6
to Pin 2 of U1), it would reduce the gain of the integrator and could
produce the symptom you are having.

Best regards,

Charles


Message: 3
Date: Wed, 09 Jul 2014 14:29:40 -0400
From: Charles Steinmetz csteinmetz@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 5200A repair
Message-ID: 20140709222949.TnBeU5B8@smtp2o.mail.yandex.net
Content-Type: text/plain; charset="us-ascii"; format=flowed

Oops!  the divider ratio of R8/R9 is 2100, not 210.

Best regards,

Charles

Message: 4
Date: Thu, 10 Jul 2014 18:25:32 +1000
From: Gerd admin@controlelectronics.com.au
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: [volt-nuts]  Fluke 5200A repair
Message-ID: 53BE4DFC.7090902@controlelectronics.com.au
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

Hello Ken,

The offset adjustment is about +-7mV to the input of U1. The amplifier has a
gain of 3 or 30 so adjustment is small. Is the problem the same on both the
10V and 100V ranges? Is the amplifier producing the full output voltage? Is
it distortion free?
Are both the 190V regulators working correctly?

Regards


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com

Hi Gerd and Charles, Thanks for the input it has given me other things to think about. I checked the output of pin 6 of U1 and I get a voltage variation of 10.2 mv from +15 to -15v adjustment on R6, so that appears to be working correctly. Now the fun starts :- using my HP 3478A to measure the dc on the output terminals I get a reading of 0.9mv on the 1 volt range, 0.3 mv on the 10 volt range, and 1.8 mv on the 100volt range. As a quick check of wether C8 may be leaky I pulled Q38 out of it's socket and got a constant DC out on the output terminals of 0.22mv on all three ranges, so maybe C8 is leaky? Using a fluke 77 3.5 digit multimeter yields vastly different DC readings, up to 5.5 v on the 100 volt range, so I suspect the Fluke 77 is reading some of the AC component on the output terminals even though DC is selected. Last night I did check the input and output on the amplifier, and on the 1 v range I was getting an input of 1v and an output of 3 volts, so the gain was correct, I did not check it on the 100v range though to see if the gain was 30. The input and output was a good clean sine wave on the amplifier. I have previously checked the 190 volt + and - supplies and they were spot on, but I will check them again before I pull the amplifier board apart again to remove C1 and C8 to test with a higher voltage other than the battery voltage of the multimeter, to see if they may be leaking at a higher voltage. Thanks, Ken ----- Message: 2 Date: Wed, 09 Jul 2014 14:21:13 -0400 From: Charles Steinmetz <csteinmetz@yandex.com> To: Discussion of precise voltage measurement <volt-nuts@febo.com> Subject: Re: [volt-nuts] Fluke 5200A repair Message-ID: <20140709222123.LMBGV5Je@smtp2o.mail.yandex.net> Content-Type: text/plain; charset="us-ascii"; format=flowed Ken wrote: >R6 slider to U1 varies from +15v to -15v but the output of U1 only >varies by mv's, I put this down to the loading of R9 to earth on the >input to pin3 of U1 (221 ohm),only allowing a small variation to the >input on pin >3 of U1, maybe this needs further investigation! Look at the circuit. The 100k potentiometer (R6) feeds a voltage divider composed of R8 and R9 (464k and 221 ohms, respectively). The tap of this divider feeds the reference input of U1. R8 and R9 attenuate the +/- 15v range of R6 by a factor of 210, so the divider tap voltage (i.e., the reference voltage for U1) only varies +/- 7.14 mV from ground for the full range of R6. U1 is an integrator, so it has lots of gain at DC (Q38 switches the integrating capacitor to give two time constants). If one of these capacitors is leaky or shorted (or there is another leakage path from Pin 6 to Pin 2 of U1), it would reduce the gain of the integrator and could produce the symptom you are having. Best regards, Charles ------------------------------ Message: 3 Date: Wed, 09 Jul 2014 14:29:40 -0400 From: Charles Steinmetz <csteinmetz@yandex.com> To: Discussion of precise voltage measurement <volt-nuts@febo.com> Subject: Re: [volt-nuts] Fluke 5200A repair Message-ID: <20140709222949.TnBeU5B8@smtp2o.mail.yandex.net> Content-Type: text/plain; charset="us-ascii"; format=flowed Oops! the divider ratio of R8/R9 is 2100, not 210. Best regards, Charles > Message: 4 Date: Thu, 10 Jul 2014 18:25:32 +1000 From: Gerd <admin@controlelectronics.com.au> To: Discussion of precise voltage measurement <volt-nuts@febo.com> Subject: [volt-nuts] Fluke 5200A repair Message-ID: <53BE4DFC.7090902@controlelectronics.com.au> Content-Type: text/plain; charset=ISO-8859-1; format=flowed Hello Ken, The offset adjustment is about +-7mV to the input of U1. The amplifier has a gain of 3 or 30 so adjustment is small. Is the problem the same on both the 10V and 100V ranges? Is the amplifier producing the full output voltage? Is it distortion free? Are both the 190V regulators working correctly? Regards --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com
G
Gerd
Fri, Jul 11, 2014 2:38 AM

Hello Ken,

Which board are you looking at? On the amplifier board (A7) C24 is the
input cap. Where is C8?
Don't rely on the multimeter to filter out the AC on the DC ranges, it's
best to use a CRO. How is the signal coming into the amplifier from the
oscillator? If the input to the amplifier is disconnected, is the offset
still there?

Regards
On 10/07/14 20:42, Kgoodhew wrote:

Hi Gerd and Charles,
Thanks for the input it has given me other things to think about.
I checked the output of pin 6 of U1 and I get a voltage variation of 10.2 mv
from +15 to -15v adjustment on R6, so that appears to be working correctly.
Now the fun starts :- using my HP 3478A to measure the dc on the output
terminals I get  a reading of 0.9mv on the 1 volt range, 0.3 mv on the 10
volt range, and 1.8 mv on the 100volt range.
As a quick check of wether C8 may be leaky I pulled Q38 out of it's socket
and got a constant DC out on the output terminals of 0.22mv on all three
ranges, so maybe C8 is leaky?
Using a fluke 77 3.5 digit multimeter yields vastly different DC readings,
up to 5.5 v on the 100 volt range, so I suspect the Fluke 77 is reading some
of the AC component on the output terminals even though DC is selected.
Last night I did check the input and output on the amplifier, and on the 1 v
range I was getting an input of 1v and an output of 3 volts, so the gain was
correct, I did not check it on the 100v range though to see if the gain was
30.
The input and output was a good clean sine wave on the amplifier.
I have previously checked the 190 volt + and - supplies and they were spot
on, but I will check them again before I pull the amplifier board apart
again to remove C1 and C8 to test with a higher voltage other than the
battery voltage of the multimeter, to see if they may be leaking at a higher
voltage.
Thanks,
Ken


Message: 2
Date: Wed, 09 Jul 2014 14:21:13 -0400
From: Charles Steinmetz csteinmetz@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 5200A repair
Message-ID: 20140709222123.LMBGV5Je@smtp2o.mail.yandex.net
Content-Type: text/plain; charset="us-ascii"; format=flowed

Ken wrote:

R6 slider to U1 varies from +15v to -15v but the output of U1 only
varies by mv's, I put this down to the loading of R9 to earth on the
input to pin3 of U1 (221 ohm),only allowing a small variation to the
input on pin
3 of U1, maybe this needs further investigation!

Look at the circuit.  The 100k potentiometer (R6) feeds a voltage divider
composed of R8 and R9 (464k and 221 ohms, respectively).  The tap of this
divider feeds the reference input of U1.  R8 and R9 attenuate the +/- 15v
range of R6 by a factor of 210, so the divider tap voltage (i.e., the
reference voltage for U1) only varies +/- 7.14 mV from ground for the full
range of R6.

U1 is an integrator, so it has lots of gain at DC (Q38 switches the
integrating capacitor to give two time constants).  If one of these
capacitors is leaky or shorted (or there is another leakage path from Pin 6
to Pin 2 of U1), it would reduce the gain of the integrator and could
produce the symptom you are having.

Best regards,

Charles


Message: 3
Date: Wed, 09 Jul 2014 14:29:40 -0400
From: Charles Steinmetz csteinmetz@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 5200A repair
Message-ID: 20140709222949.TnBeU5B8@smtp2o.mail.yandex.net
Content-Type: text/plain; charset="us-ascii"; format=flowed

Oops!  the divider ratio of R8/R9 is 2100, not 210.

Best regards,

Charles

Message: 4
Date: Thu, 10 Jul 2014 18:25:32 +1000
From: Gerd admin@controlelectronics.com.au
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: [volt-nuts]  Fluke 5200A repair
Message-ID: 53BE4DFC.7090902@controlelectronics.com.au
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

Hello Ken,

The offset adjustment is about +-7mV to the input of U1. The amplifier has a
gain of 3 or 30 so adjustment is small. Is the problem the same on both the
10V and 100V ranges? Is the amplifier producing the full output voltage? Is
it distortion free?
Are both the 190V regulators working correctly?

Regards


This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
sig.html Control Electronics Pty Ltd
The home of Shiftmaster
Quality used test equipment
Phone: 02 9456 0736
Website: www.controlelectronics.com.au
http://www.controlelectronics.com.au/*
Email*: admin@controlelectronics.com.au
mailto:admin@controlelectronics.com.au
ABN 86 109 858 022

The email addresses in this message are not to be added to any database,
list, or mailing list without my prior explicit written permission. This
email and any files transmitted with it are confidential and intended
solely for the use of the individual or entity to whom they are
addressed. If you have received this email in error please notify the
system manager. Please note that any views or opinions presented in this
email are solely those of the author and do not necessarily represent
those of the company. Finally, the recipient should check this email and
any attachments for the presence of viruses. The company accepts no
liability for any damage caused by any virus transmitted by this email.

Hello Ken, Which board are you looking at? On the amplifier board (A7) C24 is the input cap. Where is C8? Don't rely on the multimeter to filter out the AC on the DC ranges, it's best to use a CRO. How is the signal coming into the amplifier from the oscillator? If the input to the amplifier is disconnected, is the offset still there? Regards On 10/07/14 20:42, Kgoodhew wrote: > Hi Gerd and Charles, > Thanks for the input it has given me other things to think about. > I checked the output of pin 6 of U1 and I get a voltage variation of 10.2 mv > from +15 to -15v adjustment on R6, so that appears to be working correctly. > Now the fun starts :- using my HP 3478A to measure the dc on the output > terminals I get a reading of 0.9mv on the 1 volt range, 0.3 mv on the 10 > volt range, and 1.8 mv on the 100volt range. > As a quick check of wether C8 may be leaky I pulled Q38 out of it's socket > and got a constant DC out on the output terminals of 0.22mv on all three > ranges, so maybe C8 is leaky? > Using a fluke 77 3.5 digit multimeter yields vastly different DC readings, > up to 5.5 v on the 100 volt range, so I suspect the Fluke 77 is reading some > of the AC component on the output terminals even though DC is selected. > Last night I did check the input and output on the amplifier, and on the 1 v > range I was getting an input of 1v and an output of 3 volts, so the gain was > correct, I did not check it on the 100v range though to see if the gain was > 30. > The input and output was a good clean sine wave on the amplifier. > I have previously checked the 190 volt + and - supplies and they were spot > on, but I will check them again before I pull the amplifier board apart > again to remove C1 and C8 to test with a higher voltage other than the > battery voltage of the multimeter, to see if they may be leaking at a higher > voltage. > Thanks, > Ken > > ----- > > Message: 2 > Date: Wed, 09 Jul 2014 14:21:13 -0400 > From: Charles Steinmetz <csteinmetz@yandex.com> > To: Discussion of precise voltage measurement <volt-nuts@febo.com> > Subject: Re: [volt-nuts] Fluke 5200A repair > Message-ID: <20140709222123.LMBGV5Je@smtp2o.mail.yandex.net> > Content-Type: text/plain; charset="us-ascii"; format=flowed > > Ken wrote: > >> R6 slider to U1 varies from +15v to -15v but the output of U1 only >> varies by mv's, I put this down to the loading of R9 to earth on the >> input to pin3 of U1 (221 ohm),only allowing a small variation to the >> input on pin >> 3 of U1, maybe this needs further investigation! > Look at the circuit. The 100k potentiometer (R6) feeds a voltage divider > composed of R8 and R9 (464k and 221 ohms, respectively). The tap of this > divider feeds the reference input of U1. R8 and R9 attenuate the +/- 15v > range of R6 by a factor of 210, so the divider tap voltage (i.e., the > reference voltage for U1) only varies +/- 7.14 mV from ground for the full > range of R6. > > U1 is an integrator, so it has lots of gain at DC (Q38 switches the > integrating capacitor to give two time constants). If one of these > capacitors is leaky or shorted (or there is another leakage path from Pin 6 > to Pin 2 of U1), it would reduce the gain of the integrator and could > produce the symptom you are having. > > Best regards, > > Charles > > > > > > ------------------------------ > > Message: 3 > Date: Wed, 09 Jul 2014 14:29:40 -0400 > From: Charles Steinmetz <csteinmetz@yandex.com> > To: Discussion of precise voltage measurement <volt-nuts@febo.com> > Subject: Re: [volt-nuts] Fluke 5200A repair > Message-ID: <20140709222949.TnBeU5B8@smtp2o.mail.yandex.net> > Content-Type: text/plain; charset="us-ascii"; format=flowed > > Oops! the divider ratio of R8/R9 is 2100, not 210. > > Best regards, > > Charles > > > > Message: 4 > Date: Thu, 10 Jul 2014 18:25:32 +1000 > From: Gerd <admin@controlelectronics.com.au> > To: Discussion of precise voltage measurement <volt-nuts@febo.com> > Subject: [volt-nuts] Fluke 5200A repair > Message-ID: <53BE4DFC.7090902@controlelectronics.com.au> > Content-Type: text/plain; charset=ISO-8859-1; format=flowed > > Hello Ken, > > The offset adjustment is about +-7mV to the input of U1. The amplifier has a > gain of 3 or 30 so adjustment is small. Is the problem the same on both the > 10V and 100V ranges? Is the amplifier producing the full output voltage? Is > it distortion free? > Are both the 190V regulators working correctly? > > Regards > > > > --- > This email is free from viruses and malware because avast! Antivirus protection is active. > http://www.avast.com > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > -- sig.html *Control Electronics Pty Ltd* The home of Shiftmaster Quality used test equipment *Phone*: 02 9456 0736 *Website*: www.controlelectronics.com.au <http://www.controlelectronics.com.au/>* Email*: admin@controlelectronics.com.au <mailto:admin@controlelectronics.com.au> ABN 86 109 858 022 ------------------------------------------------------------------------ The email addresses in this message are not to be added to any database, list, or mailing list without my prior explicit written permission. This email and any files transmitted with it are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you have received this email in error please notify the system manager. Please note that any views or opinions presented in this email are solely those of the author and do not necessarily represent those of the company. Finally, the recipient should check this email and any attachments for the presence of viruses. The company accepts no liability for any damage caused by any virus transmitted by this email.