volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Fluke 5200A instability...

CH
Chuck Harris
Wed, Jul 29, 2015 3:35 AM

Before I spend a lot more time on it, does anyone have any experience
with the Fluke 5200A's feedback loop oscillating when in the 100Hz band
position?

Mine started to oscillate at about a 1/2Hz rate a while back. It is two
to three orders of magnitude better when in the 1KHz and higher band
positions.

I can tell it is oscillating because if I watch the 3 least significant
digits on my 3456A voltmeter, in the AC position, they keep repeating
over and over and over... sort of fits a 1-2-3-1-2-3... count.

I have replaced the bad capacitors in the power supply, and on
a few of the boards.  I have checked the supplies for voltage and
ripple, and AFAIK they are ok.

I tested the 2uf mylar integrator cap that is employed while in the
100Hz band, and replaced more than a few carbon composition resistors
that were between 20 and 50% out of tolerance with no apparent affect.

The carbon composition resistors I replaced were used to digitally
adjust the gain of the integrator, so seemed likely to have an effect
on the stability criterion of the feedback loop..

The DC Reference voltage seems to be very stable, and can be selected
up and down with the voltage selection switches.

I have tested all of the tantalums for shorts, and found none.

I am running out of ideas.

TIA,

-Chuck Harris

Before I spend a lot more time on it, does anyone have any experience with the Fluke 5200A's feedback loop oscillating when in the 100Hz band position? Mine started to oscillate at about a 1/2Hz rate a while back. It is two to three orders of magnitude better when in the 1KHz and higher band positions. I can tell it is oscillating because if I watch the 3 least significant digits on my 3456A voltmeter, in the AC position, they keep repeating over and over and over... sort of fits a 1-2-3-1-2-3... count. I have replaced the bad capacitors in the power supply, and on a few of the boards. I have checked the supplies for voltage and ripple, and AFAIK they are ok. I tested the 2uf mylar integrator cap that is employed while in the 100Hz band, and replaced more than a few carbon composition resistors that were between 20 and 50% out of tolerance with no apparent affect. The carbon composition resistors I replaced were used to digitally adjust the gain of the integrator, so seemed likely to have an effect on the stability criterion of the feedback loop.. The DC Reference voltage seems to be very stable, and can be selected up and down with the voltage selection switches. I have tested all of the tantalums for shorts, and found none. I am running out of ideas. TIA, -Chuck Harris
TM
Tom Miller
Wed, Jul 29, 2015 7:12 AM

How about any aluminum electrolytics that might be used for coupling
capacitors? I had a problem with a DC reference that had some aluminum
electrolytic caps that were leaky and caused feedback problems.

If memory holds, they were in the 10 to 30 uF range.

Just a thought.

Tom

----- Original Message -----
From: "Chuck Harris" cfharris@erols.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Tuesday, July 28, 2015 11:35 PM
Subject: [volt-nuts] Fluke 5200A instability...

Before I spend a lot more time on it, does anyone have any experience
with the Fluke 5200A's feedback loop oscillating when in the 100Hz band
position?

Mine started to oscillate at about a 1/2Hz rate a while back. It is two
to three orders of magnitude better when in the 1KHz and higher band
positions.

I can tell it is oscillating because if I watch the 3 least significant
digits on my 3456A voltmeter, in the AC position, they keep repeating
over and over and over... sort of fits a 1-2-3-1-2-3... count.

I have replaced the bad capacitors in the power supply, and on
a few of the boards.  I have checked the supplies for voltage and
ripple, and AFAIK they are ok.

I tested the 2uf mylar integrator cap that is employed while in the
100Hz band, and replaced more than a few carbon composition resistors
that were between 20 and 50% out of tolerance with no apparent affect.

The carbon composition resistors I replaced were used to digitally
adjust the gain of the integrator, so seemed likely to have an effect
on the stability criterion of the feedback loop..

The DC Reference voltage seems to be very stable, and can be selected
up and down with the voltage selection switches.

I have tested all of the tantalums for shorts, and found none.

I am running out of ideas.

TIA,

-Chuck Harris


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

How about any aluminum electrolytics that might be used for coupling capacitors? I had a problem with a DC reference that had some aluminum electrolytic caps that were leaky and caused feedback problems. If memory holds, they were in the 10 to 30 uF range. Just a thought. Tom ----- Original Message ----- From: "Chuck Harris" <cfharris@erols.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Tuesday, July 28, 2015 11:35 PM Subject: [volt-nuts] Fluke 5200A instability... > Before I spend a lot more time on it, does anyone have any experience > with the Fluke 5200A's feedback loop oscillating when in the 100Hz band > position? > > Mine started to oscillate at about a 1/2Hz rate a while back. It is two > to three orders of magnitude better when in the 1KHz and higher band > positions. > > I can tell it is oscillating because if I watch the 3 least significant > digits on my 3456A voltmeter, in the AC position, they keep repeating > over and over and over... sort of fits a 1-2-3-1-2-3... count. > > I have replaced the bad capacitors in the power supply, and on > a few of the boards. I have checked the supplies for voltage and > ripple, and AFAIK they are ok. > > I tested the 2uf mylar integrator cap that is employed while in the > 100Hz band, and replaced more than a few carbon composition resistors > that were between 20 and 50% out of tolerance with no apparent affect. > > The carbon composition resistors I replaced were used to digitally > adjust the gain of the integrator, so seemed likely to have an effect > on the stability criterion of the feedback loop.. > > The DC Reference voltage seems to be very stable, and can be selected > up and down with the voltage selection switches. > > I have tested all of the tantalums for shorts, and found none. > > I am running out of ideas. > > TIA, > > -Chuck Harris > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
CH
Chuck Harris
Wed, Jul 29, 2015 1:27 PM

Hi Tom,

That's an idea, but they didn't use any.  They used a lot of tantalums...
a mix of the orange drop style, and the hermetically sealed axial leaded
variety.  Some are operated disgustingly close to the voltage ratings...
and some are used in that suspicious back-to-back way of making an NP
tantalum cap.... but no aluminum electrolytics are used anywhere in the
signal paths... just tantalum, mylar, and silver mica.

But, I will look again...

As always, I found that the half dozen, or so, Mexican made electrolytics
were very bad, so they went away.  They were used in the power supply,
and to slow down some TTL gates that drive relays.

I have to wonder why a company that routinely used high quality Sprague
tantalums, and electrolytics would toss in a few of these Mexican timebombs.

-Chuck Harris

Tom Miller wrote:

How about any aluminum electrolytics that might be used for coupling capacitors? I
had a problem with a DC reference that had some aluminum electrolytic caps that were
leaky and caused feedback problems.

If memory holds, they were in the 10 to 30 uF range.

Just a thought.

Tom

Hi Tom, That's an idea, but they didn't use any. They used a lot of tantalums... a mix of the orange drop style, and the hermetically sealed axial leaded variety. Some are operated disgustingly close to the voltage ratings... and some are used in that suspicious back-to-back way of making an NP tantalum cap.... but no aluminum electrolytics are used anywhere in the signal paths... just tantalum, mylar, and silver mica. But, I will look again... As always, I found that the half dozen, or so, Mexican made electrolytics were very bad, so they went away. They were used in the power supply, and to slow down some TTL gates that drive relays. I have to wonder why a company that routinely used high quality Sprague tantalums, and electrolytics would toss in a few of these Mexican timebombs. -Chuck Harris Tom Miller wrote: > How about any aluminum electrolytics that might be used for coupling capacitors? I > had a problem with a DC reference that had some aluminum electrolytic caps that were > leaky and caused feedback problems. > > If memory holds, they were in the 10 to 30 uF range. > > Just a thought. > > Tom
CH
Chuck Harris
Thu, Aug 27, 2015 4:40 PM

Back in July, I posted this note about a stability problem I
was having with my Fluke 5200A.  I didn't get much response,
so I dove in with both feet.

The symptom was viewed on a trio of HP3456A 6-1/2 digit DVM's,
when measuring the 10V 60Hz output of a Fluke 5200A.

The 3456A's showed a 20mv offset, and an instability in the
tree least significant digits.  The instability was cyclical,
with about a 1/3Hz period.  The instability was gone when
the range switch was changed from 100Hz to 1KHz.  It clearly
diminished as the frequency rose farther and farther above
60Hz.

My first stop was the power supply, and checking and replacing
about 1/2 of the electrolytic capacitors.  They had ESR's on
the high side of good, but still seemed to be functioning.

Result:  No noticeable difference.

The next stop was to check all shields, grounds, isolation
between various guards and isolated grounds, etc...

Result, nothing found, no noticeable difference.

I then went to the AC/DC board, which housed an AC to DC
rectifier, various feedback paths from the sense terminals,
the integrator, and the control amplifier... It should be
accountable for any feedback irregularities.

I found numerous carbon composition resistors that were between
20% and 50% high of their intended values, and replaced them.

I checked, and found the 2uf mylar 100Hz range integrator
capacitor had some leakage, and a bit of dielectric absorption,
so I replaced it.

Tested all the tantalums, they were good...

Result, no noticeable difference... Hmmm?

I took a little time out to make an extender board using my
PCB mill.  I couldn't find the appropriate connector, so I
cut the middle out of a 100 pin 0.1" spacing card edge connector,
and mounted the two halves to the extender.

I tested the reference board, the oscillator, and the oscillator
control, and found minor offset issues, a pernicious bit of
digital ripple on the -15V(f) supply... that seemed like it should,
but didn't really matter... and a couple of more carbon composition
resistors that were well out of tolerance.

Result, no noticeable difference... Hmmm?  Hmmm?

I checked out the power amplifier board, and found several
electrolytics that were high ESR, loose screws on the output
transistor's heatsink.  Offset in need of adjustment, but
nothing much else.

Result, no noticeable difference... What's with this thing?

After concluding that the instability had something to do with
the 100Hz range, and checking all of the circuitry associated
with the 100Hz range selection signals, and finding nothing,
I finally did an experiment I should have done before:

I made some test measurements at 110Hz on the 100Hz range, and
then made the identical test measurements at 0.110KHz on the
1KHz range, and.....

They were exactly the same!  I can rule out anything to do with
the 100Hz range selection, and just look at the output frequency
as being important to the failure.

I was puzzled by how I was supposed to measure mv signals on
top of a 10V 60Hz sinewave, and got the vague idea to use my
dusty old highly neglected Fluke 893A AC/DC differential
voltmeter.  I figured that it could show the instability in the
galvanometer meter used for nulling the bridge....

On trying, I found that I could not see the instability on the
893A's null meter, and further, the Fluke 893A agreed with the
Fluke 5200 calibrator as to the AC values being measured....
everywhere, every voltage, and throughout the differential
voltmeter's frequency range!

This can't be a coincidence.  A differential voltmeter uses
a completely different method of measuring AC and DC from a
DVM, and it says the 5200A is working properly.

So, on advice from my friend Tom B., I set my HP3336B synthesizer
up to output 60Hz, and 7dBm, and plugged that into the HP3456A
DVM, and there it was:  The DVM can read voltages at exactly
60Hz pretty well, but fractionally higher, or lower, it shows
a beat note in the reading... exactly what I was seeing on the
Fluke 5200A.

Having had problems with HP3456A power supplies in the past, I
did a quick check on its filters, and found them all to be just
fine...

Doing something that comes hard for me, I pulled out the
instruction manual for the 3456A, and read through the operating
instructions.

The HP3456A is not specified to operate below 400Hz AC unless
the filter button (that I never use) on the front panel is pressed.

I pressed the filter button, and the meter fell into line, and
read a consistent, and fairly stable value.

Wow!  Such a ride to take for a simple case of operator error!

An additional bit of insight was garnered, when I read the AC
specs section of the HP3456A, and tried to understand them:

6-1/2 digit (>=1PLC)
Filter OFF:
400Hz-20KHz: ±(0.07% + 730)

Filter ON:
10Hz-30Hz: ±(0.47% + 450)
30Hz-20kHz: ±(0.07% + 730)

Where PLC = Integrator's Power Line Cycles

Given a 10.0000V AC 60Hz signal, the HP 3456A could read it
as anywhere between 10.1430V to 9.8570V, and still be within
specifications.  Further, it could count wildly in that range,
and still be in specification.

Out comes the Fluke 540B, and a recalibration for my poor
maligned 5200A.

On the bright side, it was full of parts that had degraded over
time... not that it seemed to matter...

An so it goes...

-Chuck Harris

Back in July, I posted this note about a stability problem I was having with my Fluke 5200A. I didn't get much response, so I dove in with both feet. The symptom was viewed on a trio of HP3456A 6-1/2 digit DVM's, when measuring the 10V 60Hz output of a Fluke 5200A. The 3456A's showed a 20mv offset, and an instability in the tree least significant digits. The instability was cyclical, with about a 1/3Hz period. The instability was gone when the range switch was changed from 100Hz to 1KHz. It clearly diminished as the frequency rose farther and farther above 60Hz. My first stop was the power supply, and checking and replacing about 1/2 of the electrolytic capacitors. They had ESR's on the high side of good, but still seemed to be functioning. Result: No noticeable difference. The next stop was to check all shields, grounds, isolation between various guards and isolated grounds, etc... Result, nothing found, no noticeable difference. I then went to the AC/DC board, which housed an AC to DC rectifier, various feedback paths from the sense terminals, the integrator, and the control amplifier... It should be accountable for any feedback irregularities. I found numerous carbon composition resistors that were between 20% and 50% high of their intended values, and replaced them. I checked, and found the 2uf mylar 100Hz range integrator capacitor had some leakage, and a bit of dielectric absorption, so I replaced it. Tested all the tantalums, they were good... Result, no noticeable difference... Hmmm? I took a little time out to make an extender board using my PCB mill. I couldn't find the appropriate connector, so I cut the middle out of a 100 pin 0.1" spacing card edge connector, and mounted the two halves to the extender. I tested the reference board, the oscillator, and the oscillator control, and found minor offset issues, a pernicious bit of digital ripple on the -15V(f) supply... that seemed like it should, but didn't really matter... and a couple of more carbon composition resistors that were well out of tolerance. Result, no noticeable difference... Hmmm? Hmmm? I checked out the power amplifier board, and found several electrolytics that were high ESR, loose screws on the output transistor's heatsink. Offset in need of adjustment, but nothing much else. Result, no noticeable difference... What's with this thing? After concluding that the instability had something to do with the 100Hz range, and checking all of the circuitry associated with the 100Hz range selection signals, and finding nothing, I finally did an experiment I should have done before: I made some test measurements at 110Hz on the 100Hz range, and then made the identical test measurements at 0.110KHz on the 1KHz range, and..... They were exactly the same! I can rule out anything to do with the 100Hz range selection, and just look at the output frequency as being important to the failure. I was puzzled by how I was supposed to measure mv signals on top of a 10V 60Hz sinewave, and got the vague idea to use my dusty old highly neglected Fluke 893A AC/DC differential voltmeter. I figured that it could show the instability in the galvanometer meter used for nulling the bridge.... On trying, I found that I could not see the instability on the 893A's null meter, and further, the Fluke 893A agreed with the Fluke 5200 calibrator as to the AC values being measured.... everywhere, every voltage, and throughout the differential voltmeter's frequency range! This can't be a coincidence. A differential voltmeter uses a completely different method of measuring AC and DC from a DVM, and it says the 5200A is working properly. So, on advice from my friend Tom B., I set my HP3336B synthesizer up to output 60Hz, and 7dBm, and plugged that into the HP3456A DVM, and there it was: The DVM can read voltages at exactly 60Hz pretty well, but fractionally higher, or lower, it shows a beat note in the reading... exactly what I was seeing on the Fluke 5200A. Having had problems with HP3456A power supplies in the past, I did a quick check on its filters, and found them all to be just fine... Doing something that comes hard for me, I pulled out the instruction manual for the 3456A, and read through the operating instructions. The HP3456A is not specified to operate below 400Hz AC unless the filter button (that I never use) on the front panel is pressed. I pressed the filter button, and the meter fell into line, and read a consistent, and fairly stable value. Wow! Such a ride to take for a simple case of operator error! An additional bit of insight was garnered, when I read the AC specs section of the HP3456A, and tried to understand them: 6-1/2 digit (>=1PLC) Filter OFF: 400Hz-20KHz: ±(0.07% + 730) Filter ON: 10Hz-30Hz: ±(0.47% + 450) 30Hz-20kHz: ±(0.07% + 730) Where PLC = Integrator's Power Line Cycles Given a 10.0000V AC 60Hz signal, the HP 3456A could read it as anywhere between 10.1430V to 9.8570V, and still be within specifications. Further, it could count wildly in that range, and still be in specification. Out comes the Fluke 540B, and a recalibration for my poor maligned 5200A. On the bright side, it was full of parts that had degraded over time... not that it seemed to matter... An so it goes... -Chuck Harris
DC
David C. Partridge
Thu, Aug 27, 2015 5:35 PM

Chuck I so feel for you one this one.  I almost went down the exact same
rabbit hole about two years back when I didn't set the AC Filter to ON on my
7081.

Luckily I didn't get very far down the "my 5200A isn't working" track before
I was reminded by a friend that I needed to turn on the filter for <400Hz
signals.

Cheers
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Chuck
Harris
Sent: 27 August 2015 17:41
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 5200A instability... A resolution...

:

Doing something that comes hard for me, I pulled out the instruction manual
for the 3456A, and read through the operating instructions.

The HP3456A is not specified to operate below 400Hz AC unless the filter
button (that I never use) on the front panel is pressed.

I pressed the filter button, and the meter fell into line, and read a
consistent, and fairly stable value.

Chuck I so feel for you one this one. I almost went down the exact same rabbit hole about two years back when I didn't set the AC Filter to ON on my 7081. Luckily I didn't get very far down the "my 5200A isn't working" track before I was reminded by a friend that I needed to turn on the filter for <400Hz signals. Cheers Dave -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Chuck Harris Sent: 27 August 2015 17:41 To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A instability... A resolution... : Doing something that comes hard for me, I pulled out the instruction manual for the 3456A, and read through the operating instructions. The HP3456A is not specified to operate below 400Hz AC unless the filter button (that I never use) on the front panel is pressed. I pressed the filter button, and the meter fell into line, and read a consistent, and fairly stable value.
CH
Chuck Harris
Thu, Aug 27, 2015 5:50 PM

Hi David,

I don't mind so much the journey, as I do being disappointed
in the performance of the HP3456A's AC section.  DC and Ohms,
it is a wonderful DVM.  AC, not so much.

Since it times its integrators to the power line frequency, it
is dooming itself to work stably only at those exact frequencies.

Wouldn't it have been better to arrange things so that the
integrator randomly frequency hopped about a range of frequencies?

With computer control, it should be fairly easy to compensate for
the differing integration times such a scheme would require...
even back then.

-Chuck Harris

David C. Partridge wrote:

Chuck I so feel for you one this one.  I almost went down the exact same
rabbit hole about two years back when I didn't set the AC Filter to ON on my
7081.

Luckily I didn't get very far down the "my 5200A isn't working" track before
I was reminded by a friend that I needed to turn on the filter for <400Hz
signals.

Cheers
Dave

Hi David, I don't mind so much the journey, as I do being disappointed in the performance of the HP3456A's AC section. DC and Ohms, it is a wonderful DVM. AC, not so much. Since it times its integrators to the power line frequency, it is dooming itself to work stably only at those exact frequencies. Wouldn't it have been better to arrange things so that the integrator randomly frequency hopped about a range of frequencies? With computer control, it should be fairly easy to compensate for the differing integration times such a scheme would require... even back then. -Chuck Harris David C. Partridge wrote: > Chuck I so feel for you one this one. I almost went down the exact same > rabbit hole about two years back when I didn't set the AC Filter to ON on my > 7081. > > Luckily I didn't get very far down the "my 5200A isn't working" track before > I was reminded by a friend that I needed to turn on the filter for <400Hz > signals. > > Cheers > Dave >
JP
John Phillips
Thu, Aug 27, 2015 9:15 PM

The HP 3455A had a much better AC section but it was a pain to cal.

HP work hard to make cal easy and settled for a weak AC converter so they
could do a closed case cal.

The 3455A, 3457A and 3458A all integrate at line frequency and all have a
better AC converter.

Integration is tied to line frequency to cancel line noise not to have the
interrogator more accurate at what ever integration frequency.

Integration time is also used for all the measurements not just the AC
converter.

3457A is a better meter than the 3456A but people do not seem to respect
it. It may be the display but it is the first of the HP meters to do stats.
You may be able to find a 3457A for about or less money than a 3456A It
will be a lot easier to carry as well.

On Thu, Aug 27, 2015 at 10:50 AM, Chuck Harris cfharris@erols.com wrote:

Hi David,

I don't mind so much the journey, as I do being disappointed
in the performance of the HP3456A's AC section.  DC and Ohms,
it is a wonderful DVM.  AC, not so much.

Since it times its integrators to the power line frequency, it
is dooming itself to work stably only at those exact frequencies.

Wouldn't it have been better to arrange things so that the
integrator randomly frequency hopped about a range of frequencies?

With computer control, it should be fairly easy to compensate for
the differing integration times such a scheme would require...
even back then.

-Chuck Harris

David C. Partridge wrote:

Chuck I so feel for you one this one.  I almost went down the exact same
rabbit hole about two years back when I didn't set the AC Filter to ON on
my
7081.

Luckily I didn't get very far down the "my 5200A isn't working" track
before
I was reminded by a friend that I needed to turn on the filter for <400Hz
signals.

Cheers
Dave


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--

John Phillips

The HP 3455A had a much better AC section but it was a pain to cal. HP work hard to make cal easy and settled for a weak AC converter so they could do a closed case cal. The 3455A, 3457A and 3458A all integrate at line frequency and all have a better AC converter. Integration is tied to line frequency to cancel line noise not to have the interrogator more accurate at what ever integration frequency. Integration time is also used for all the measurements not just the AC converter. 3457A is a better meter than the 3456A but people do not seem to respect it. It may be the display but it is the first of the HP meters to do stats. You may be able to find a 3457A for about or less money than a 3456A It will be a lot easier to carry as well. On Thu, Aug 27, 2015 at 10:50 AM, Chuck Harris <cfharris@erols.com> wrote: > Hi David, > > I don't mind so much the journey, as I do being disappointed > in the performance of the HP3456A's AC section. DC and Ohms, > it is a wonderful DVM. AC, not so much. > > Since it times its integrators to the power line frequency, it > is dooming itself to work stably only at those exact frequencies. > > Wouldn't it have been better to arrange things so that the > integrator randomly frequency hopped about a range of frequencies? > > With computer control, it should be fairly easy to compensate for > the differing integration times such a scheme would require... > even back then. > > -Chuck Harris > > > David C. Partridge wrote: > >> Chuck I so feel for you one this one. I almost went down the exact same >> rabbit hole about two years back when I didn't set the AC Filter to ON on >> my >> 7081. >> >> Luckily I didn't get very far down the "my 5200A isn't working" track >> before >> I was reminded by a friend that I needed to turn on the filter for <400Hz >> signals. >> >> Cheers >> Dave >> >> _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- *John Phillips*