Pete,
At 4:58 AM +0000 12/27/08, time-nuts-request@febo.com wrote:
Message: 2
Date: Fri, 26 Dec 2008 19:10:47 -0700
From: "Pete" peterawson@earthlink.net
Subject: Re: [time-nuts] Sub Pico Second Phase logger
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Joe,
I have been following this thread from the start & learned a lot from it.
Thanks to you and Bruce for the insights.
I thought I would take a look at a simple, passive approach, just to
see what happens. I also limited my attention to stable 5 or 10 MHz
sources & observation windows of <100s. I used a Mini-circuits
SYPD-1 PD with a PSCQ-2-10.5 90 deg hybrid plus an ATM
P1506 phase shifter (delta=0.32ns). There are lots of other items
which will work just as well, this stuff happens to be on hand.
The SYPD-1 also needs a diplexer to terminate the high order
products & low pass filter the desired output; this is a cobble
together item from the parts box. I used a PC logging program
with my HP 3478A to take DC readings from the diplexer output.
What do you mean by "diplexer"?
Thus armed, I first checked the DVM logging with no RF in to see
if it was stable enough to yield a useable noise floor. This first result
was good enough. Reading 720 points over 100 seconds; the data
was limited to -1, 0 or +1 uV.
Then I applied 10MHz @ to the 90 deg hybrid & connected one
hybrid output to the SYPD-1 LO & the other hybrid output to the
ATM P1506. The ATM output then drives the SYPD-1 RF port.
I measured both SYPD-1 inputs to be +6.4dBm. Now I was able
to set the ATM phase to zero the SYPD-1 output & verify it was
stable over the 100s time I was interested in. The 3478A readings
ranged from -1 to +2uV during this time with STDDEV of 0.6uV
-again, good enough for me.
Finally, I used a HP 105B to get a quiet 5MHz source & a 1MHz
ext. reference for my HP3336C sythesizer. All this to give me 2
"nearly" identical. phase locked sources to measure the calibration
factor of the SYPD-1/Diplexer output. The 3336C was set to
5,000,000.005Hz & +6.4dBm driving the SYPD-1 LO port. The
105B 5MHz output was attenuated to +6.4dBm driving the RF
port. The resulting 5mHz data was logged & analyzed at the zero
crossings (max gain). The result is a phase sensitivity of 1.22uRad/uV.
The + & - zero crossings were identical and the 1.22uRad/uV factor
is linear (+/- 2 %) up to +/- 20 % of FS. FS = 1.023V
If my math is about right, @10MHz, 1uRad = 16fs; so a noise floor
of +/- 2uV (+/-2.5uRad) equals +/-40fs. I expect that, if 2 sources
are near zero phase, then it should only require <10s to estimate
the actual phase error within 1ps, or less. Anyone agree?
I await & hope for constructive comments and/or corrections.
I've done similar things, and can also get random errors in the
femtoseconds, but only over short periods of time. The key is
temperature sensitivity. I would be suspicious of the 90 degree
hybrid in particular, although the phase detector also has an
opinion. The splitter phase unbalance is also quite sensitive to
drive frequency, so the frequency stable sources are essential.
Joe