volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

HP-419A Fluke 845A Mods - shared link

RE
Randy Evans
Wed, Jun 24, 2015 3:40 AM

I have created a shared link to my working folder with the details of the
HP-419A and Fluke 845A modifications.  It is an in-process design so I
suspect there may be many questions.  Just send your questions and I will
try to answer them.  In the meantime, I will try to find time to clean up
the design a bit.

Hopefully I created the link correctly.  Let me know if you have any issues
with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG

I have created a shared link to my working folder with the details of the HP-419A and Fluke 845A modifications. It is an in-process design so I suspect there may be many questions. Just send your questions and I will try to answer them. In the meantime, I will try to find time to clean up the design a bit. Hopefully I created the link correctly. Let me know if you have any issues with the link. https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 Randy Evans AE6YG
DM
Dave M
Wed, Jun 24, 2015 5:49 PM

Rady,
Many thanks for the uploads.
One issue... the marked-up PDF of the HP 419A schematic is virtually
unreadable.  Could you post a higher resolution image of the markup details?
Don't really need the full schematic, just the areas where the connections
to the new circuit are made.

Thanks again,
Dave M

Randy Evans wrote:

I have created a shared link to my working folder with the details of
the HP-419A and Fluke 845A modifications.  It is an in-process design
so I suspect there may be many questions.  Just send your questions
and I will try to answer them.  In the meantime, I will try to find
time to clean up the design a bit.

Hopefully I created the link correctly.  Let me know if you have any
issues with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG

Rady, Many thanks for the uploads. One issue... the marked-up PDF of the HP 419A schematic is virtually unreadable. Could you post a higher resolution image of the markup details? Don't really need the full schematic, just the areas where the connections to the new circuit are made. Thanks again, Dave M Randy Evans wrote: > I have created a shared link to my working folder with the details of > the HP-419A and Fluke 845A modifications. It is an in-process design > so I suspect there may be many questions. Just send your questions > and I will try to answer them. In the meantime, I will try to find > time to clean up the design a bit. > > Hopefully I created the link correctly. Let me know if you have any > issues with the link. > > https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 > > > Randy Evans AE6YG
RE
Randy Evans
Thu, Jun 25, 2015 1:38 AM

Dave,

I should have checked it.  I will generate a higher resolution version
tomorrow.

Randy

On Wed, Jun 24, 2015 at 10:49 AM, Dave M dgminala@mediacombb.net wrote:

Rady,
Many thanks for the uploads.
One issue... the marked-up PDF of the HP 419A schematic is virtually
unreadable.  Could you post a higher resolution image of the markup
details? Don't really need the full schematic, just the areas where the
connections to the new circuit are made.

Thanks again,
Dave M

Randy Evans wrote:

I have created a shared link to my working folder with the details of
the HP-419A and Fluke 845A modifications.  It is an in-process design
so I suspect there may be many questions.  Just send your questions
and I will try to answer them.  In the meantime, I will try to find
time to clean up the design a bit.

Hopefully I created the link correctly.  Let me know if you have any
issues with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Dave, I should have checked it. I will generate a higher resolution version tomorrow. Randy On Wed, Jun 24, 2015 at 10:49 AM, Dave M <dgminala@mediacombb.net> wrote: > Rady, > Many thanks for the uploads. > One issue... the marked-up PDF of the HP 419A schematic is virtually > unreadable. Could you post a higher resolution image of the markup > details? Don't really need the full schematic, just the areas where the > connections to the new circuit are made. > > > Thanks again, > Dave M > > > Randy Evans wrote: > >> I have created a shared link to my working folder with the details of >> the HP-419A and Fluke 845A modifications. It is an in-process design >> so I suspect there may be many questions. Just send your questions >> and I will try to answer them. In the meantime, I will try to find >> time to clean up the design a bit. >> >> Hopefully I created the link correctly. Let me know if you have any >> issues with the link. >> >> https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 >> >> >> Randy Evans AE6YG >> > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
RE
Randy Evans
Thu, Jun 25, 2015 3:09 AM

Dave,

I checked the HP-419A marked up schematic and I find it very clear and
readable.  Does anyone else have problem reading the pdf file?

Thanks,

Randy Evans

On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans randyevans2688@gmail.com
wrote:

Dave,

I should have checked it.  I will generate a higher resolution version
tomorrow.

Randy

On Wed, Jun 24, 2015 at 10:49 AM, Dave M dgminala@mediacombb.net wrote:

Rady,
Many thanks for the uploads.
One issue... the marked-up PDF of the HP 419A schematic is virtually
unreadable.  Could you post a higher resolution image of the markup
details? Don't really need the full schematic, just the areas where the
connections to the new circuit are made.

Thanks again,
Dave M

Randy Evans wrote:

I have created a shared link to my working folder with the details of
the HP-419A and Fluke 845A modifications.  It is an in-process design
so I suspect there may be many questions.  Just send your questions
and I will try to answer them.  In the meantime, I will try to find
time to clean up the design a bit.

Hopefully I created the link correctly.  Let me know if you have any
issues with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Dave, I checked the HP-419A marked up schematic and I find it very clear and readable. Does anyone else have problem reading the pdf file? Thanks, Randy Evans On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans <randyevans2688@gmail.com> wrote: > Dave, > > I should have checked it. I will generate a higher resolution version > tomorrow. > > Randy > > On Wed, Jun 24, 2015 at 10:49 AM, Dave M <dgminala@mediacombb.net> wrote: > >> Rady, >> Many thanks for the uploads. >> One issue... the marked-up PDF of the HP 419A schematic is virtually >> unreadable. Could you post a higher resolution image of the markup >> details? Don't really need the full schematic, just the areas where the >> connections to the new circuit are made. >> >> >> Thanks again, >> Dave M >> >> >> Randy Evans wrote: >> >>> I have created a shared link to my working folder with the details of >>> the HP-419A and Fluke 845A modifications. It is an in-process design >>> so I suspect there may be many questions. Just send your questions >>> and I will try to answer them. In the meantime, I will try to find >>> time to clean up the design a bit. >>> >>> Hopefully I created the link correctly. Let me know if you have any >>> issues with the link. >>> >>> https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 >>> >>> >>> Randy Evans AE6YG >>> >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > >
DM
Dave M
Thu, Jun 25, 2015 4:55 PM

Randy,
False alarm.  When I downloaded the PDF again, it came through fine.  The
file that I downloaded originally was only about 75K, not the 500K file that
I got the second time.

Sorry for the alarm.

Thanks again for sharing your work.
Dave M

Randy Evans wrote:

Dave,

I checked the HP-419A marked up schematic and I find it very clear and
readable.  Does anyone else have problem reading the pdf file?

Thanks,

Randy Evans

On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans
randyevans2688@gmail.com wrote:

Dave,

I should have checked it.  I will generate a higher resolution
version tomorrow.

Randy

On Wed, Jun 24, 2015 at 10:49 AM, Dave M dgminala@mediacombb.net
wrote:

Rady,
Many thanks for the uploads.
One issue... the marked-up PDF of the HP 419A schematic is virtually
unreadable.  Could you post a higher resolution image of the markup
details? Don't really need the full schematic, just the areas where
the connections to the new circuit are made.

Thanks again,
Dave M

Randy Evans wrote:

I have created a shared link to my working folder with the details
of the HP-419A and Fluke 845A modifications.  It is an in-process
design so I suspect there may be many questions.  Just send your
questions and I will try to answer them.  In the meantime, I will
try to find time to clean up the design a bit.

Hopefully I created the link correctly.  Let me know if you have
any issues with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG

Randy, False alarm. When I downloaded the PDF again, it came through fine. The file that I downloaded originally was only about 75K, not the 500K file that I got the second time. Sorry for the alarm. Thanks again for sharing your work. Dave M Randy Evans wrote: > Dave, > > I checked the HP-419A marked up schematic and I find it very clear and > readable. Does anyone else have problem reading the pdf file? > > Thanks, > > Randy Evans > > On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans > <randyevans2688@gmail.com> wrote: > >> Dave, >> >> I should have checked it. I will generate a higher resolution >> version tomorrow. >> >> Randy >> >> On Wed, Jun 24, 2015 at 10:49 AM, Dave M <dgminala@mediacombb.net> >> wrote: >> >>> Rady, >>> Many thanks for the uploads. >>> One issue... the marked-up PDF of the HP 419A schematic is virtually >>> unreadable. Could you post a higher resolution image of the markup >>> details? Don't really need the full schematic, just the areas where >>> the connections to the new circuit are made. >>> >>> >>> Thanks again, >>> Dave M >>> >>> >>> Randy Evans wrote: >>> >>>> I have created a shared link to my working folder with the details >>>> of the HP-419A and Fluke 845A modifications. It is an in-process >>>> design so I suspect there may be many questions. Just send your >>>> questions and I will try to answer them. In the meantime, I will >>>> try to find time to clean up the design a bit. >>>> >>>> Hopefully I created the link correctly. Let me know if you have >>>> any issues with the link. >>>> >>>> https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 >>>> >>>> >>>> Randy Evans AE6YG >>>>
LL
Ludger.Lenzen@gmx.de
Fri, Jun 26, 2015 7:50 PM

For all, working on a modification of the HP-419A:
I have uploaded to my HP-419A share point some high quality schematics, I have created for the modification process.
Link:
https://www.dropbox.com/sh/czd2uc3t0uvgggz/AAC--ykVziW9rSJkdg2J_-wZa?dl=0

Important note:
I created them using PaintNET on 64-bit Win7.
It may impossible to open them with the 32-bit Win7 version because of the memory restrictions.

I uploaded following files:
HP419A optimized original schematic complete (with doc).PDN  (the original amplifier schematic with some additional notes)
HP419A optimized original schematic complete.PDN (the original amplifier schematic)
HP419A optimized original schematic (3µV mode).PDN (amplifier schematic in the 3µV mode)

Enjoy and feel free to comment.
BR
PeLuLe

 
 

Gesendet: Donnerstag, 25. Juni 2015 um 18:55 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] HP-419A Fluke 845A Mods - shared link
Randy,
False alarm. When I downloaded the PDF again, it came through fine. The
file that I downloaded originally was only about 75K, not the 500K file that
I got the second time.

Sorry for the alarm.

Thanks again for sharing your work.
Dave M

Randy Evans wrote:

Dave,

I checked the HP-419A marked up schematic and I find it very clear and
readable. Does anyone else have problem reading the pdf file?

Thanks,

Randy Evans

On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans
randyevans2688@gmail.com wrote:

Dave,

I should have checked it. I will generate a higher resolution
version tomorrow.

Randy

On Wed, Jun 24, 2015 at 10:49 AM, Dave M dgminala@mediacombb.net
wrote:

Rady,
Many thanks for the uploads.
One issue... the marked-up PDF of the HP 419A schematic is virtually
unreadable. Could you post a higher resolution image of the markup
details? Don't really need the full schematic, just the areas where
the connections to the new circuit are made.

Thanks again,
Dave M

Randy Evans wrote:

I have created a shared link to my working folder with the details
of the HP-419A and Fluke 845A modifications. It is an in-process
design so I suspect there may be many questions. Just send your
questions and I will try to answer them. In the meantime, I will
try to find time to clean up the design a bit.

Hopefully I created the link correctly. Let me know if you have
any issues with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts[https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts]
and follow the instructions there.

For all, working on a modification of the HP-419A: I have uploaded to my HP-419A share point some high quality schematics, I have created for the modification process. Link: https://www.dropbox.com/sh/czd2uc3t0uvgggz/AAC--ykVziW9rSJkdg2J_-wZa?dl=0 Important note: I created them using PaintNET on 64-bit Win7. It may impossible to open them with the 32-bit Win7 version because of the memory restrictions. I uploaded following files: HP419A optimized original schematic complete (with doc).PDN (the original amplifier schematic with some additional notes) HP419A optimized original schematic complete.PDN (the original amplifier schematic) HP419A optimized original schematic (3µV mode).PDN (amplifier schematic in the 3µV mode) Enjoy and feel free to comment. BR PeLuLe     Gesendet: Donnerstag, 25. Juni 2015 um 18:55 Uhr Von: "Dave M" <dgminala@mediacombb.net> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Betreff: Re: [volt-nuts] HP-419A Fluke 845A Mods - shared link Randy, False alarm. When I downloaded the PDF again, it came through fine. The file that I downloaded originally was only about 75K, not the 500K file that I got the second time. Sorry for the alarm. Thanks again for sharing your work. Dave M Randy Evans wrote: > Dave, > > I checked the HP-419A marked up schematic and I find it very clear and > readable. Does anyone else have problem reading the pdf file? > > Thanks, > > Randy Evans > > On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans > <randyevans2688@gmail.com> wrote: > >> Dave, >> >> I should have checked it. I will generate a higher resolution >> version tomorrow. >> >> Randy >> >> On Wed, Jun 24, 2015 at 10:49 AM, Dave M <dgminala@mediacombb.net> >> wrote: >> >>> Rady, >>> Many thanks for the uploads. >>> One issue... the marked-up PDF of the HP 419A schematic is virtually >>> unreadable. Could you post a higher resolution image of the markup >>> details? Don't really need the full schematic, just the areas where >>> the connections to the new circuit are made. >>> >>> >>> Thanks again, >>> Dave M >>> >>> >>> Randy Evans wrote: >>> >>>> I have created a shared link to my working folder with the details >>>> of the HP-419A and Fluke 845A modifications. It is an in-process >>>> design so I suspect there may be many questions. Just send your >>>> questions and I will try to answer them. In the meantime, I will >>>> try to find time to clean up the design a bit. >>>> >>>> Hopefully I created the link correctly. Let me know if you have >>>> any issues with the link. >>>> >>>> https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 >>>> >>>> >>>> Randy Evans AE6YG >>>> _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts[https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts] and follow the instructions there.
DM
Dave M
Fri, Jun 26, 2015 10:22 PM

The .PDN image files can be opened on Win XP, Win7 & Win8 with Irfanview if
you download and install the plugin package from the Irfanview web site.  I
use XP 32-bit, and the files opened just fine for me.

Cheers,
Dave M

Ludger.Lenzen@gmx.de wrote:

For all, working on a modification of the HP-419A:
I have uploaded to my HP-419A share point some high quality
schematics, I have created for the modification process.
Link:
https://www.dropbox.com/sh/czd2uc3t0uvgggz/AAC--ykVziW9rSJkdg2J_-wZa?dl=0

Important note:
I created them using PaintNET on 64-bit Win7.
It may impossible to open them with the 32-bit Win7 version because
of the memory restrictions.

I uploaded following files:
HP419A optimized original schematic complete (with doc).PDN  (the
original amplifier schematic with some additional notes)
HP419A optimized original schematic complete.PDN (the original
amplifier schematic)
HP419A optimized original schematic (3µV mode).PDN (amplifier
schematic in the 3µV mode)

Enjoy and feel free to comment.
BR
PeLuLe

Gesendet: Donnerstag, 25. Juni 2015 um 18:55 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] HP-419A Fluke 845A Mods - shared link
Randy,
False alarm. When I downloaded the PDF again, it came through fine.
The
file that I downloaded originally was only about 75K, not the 500K
file that
I got the second time.

Sorry for the alarm.

Thanks again for sharing your work.
Dave M

Randy Evans wrote:

Dave,

I checked the HP-419A marked up schematic and I find it very clear
and readable. Does anyone else have problem reading the pdf file?

Thanks,

Randy Evans

On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans
randyevans2688@gmail.com wrote:

Dave,

I should have checked it. I will generate a higher resolution
version tomorrow.

Randy

On Wed, Jun 24, 2015 at 10:49 AM, Dave M dgminala@mediacombb.net
wrote:

Rady,
Many thanks for the uploads.
One issue... the marked-up PDF of the HP 419A schematic is
virtually unreadable. Could you post a higher resolution image of
the markup details? Don't really need the full schematic, just the
areas where the connections to the new circuit are made.

Thanks again,
Dave M

Randy Evans wrote:

I have created a shared link to my working folder with the details
of the HP-419A and Fluke 845A modifications. It is an in-process
design so I suspect there may be many questions. Just send your
questions and I will try to answer them. In the meantime, I will
try to find time to clean up the design a bit.

Hopefully I created the link correctly. Let me know if you have
any issues with the link.

https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0

Randy Evans AE6YG

The .PDN image files can be opened on Win XP, Win7 & Win8 with Irfanview if you download and install the plugin package from the Irfanview web site. I use XP 32-bit, and the files opened just fine for me. Cheers, Dave M Ludger.Lenzen@gmx.de wrote: > For all, working on a modification of the HP-419A: > I have uploaded to my HP-419A share point some high quality > schematics, I have created for the modification process. > Link: > https://www.dropbox.com/sh/czd2uc3t0uvgggz/AAC--ykVziW9rSJkdg2J_-wZa?dl=0 > > Important note: > I created them using PaintNET on 64-bit Win7. > It may impossible to open them with the 32-bit Win7 version because > of the memory restrictions. > > I uploaded following files: > HP419A optimized original schematic complete (with doc).PDN (the > original amplifier schematic with some additional notes) > HP419A optimized original schematic complete.PDN (the original > amplifier schematic) > HP419A optimized original schematic (3µV mode).PDN (amplifier > schematic in the 3µV mode) > > Enjoy and feel free to comment. > BR > PeLuLe > > > > > > Gesendet: Donnerstag, 25. Juni 2015 um 18:55 Uhr > Von: "Dave M" <dgminala@mediacombb.net> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] HP-419A Fluke 845A Mods - shared link > Randy, > False alarm. When I downloaded the PDF again, it came through fine. > The > file that I downloaded originally was only about 75K, not the 500K > file that > I got the second time. > > Sorry for the alarm. > > Thanks again for sharing your work. > Dave M > > > Randy Evans wrote: >> Dave, >> >> I checked the HP-419A marked up schematic and I find it very clear >> and readable. Does anyone else have problem reading the pdf file? >> >> Thanks, >> >> Randy Evans >> >> On Wed, Jun 24, 2015 at 6:38 PM, Randy Evans >> <randyevans2688@gmail.com> wrote: >> >>> Dave, >>> >>> I should have checked it. I will generate a higher resolution >>> version tomorrow. >>> >>> Randy >>> >>> On Wed, Jun 24, 2015 at 10:49 AM, Dave M <dgminala@mediacombb.net> >>> wrote: >>> >>>> Rady, >>>> Many thanks for the uploads. >>>> One issue... the marked-up PDF of the HP 419A schematic is >>>> virtually unreadable. Could you post a higher resolution image of >>>> the markup details? Don't really need the full schematic, just the >>>> areas where the connections to the new circuit are made. >>>> >>>> >>>> Thanks again, >>>> Dave M >>>> >>>> >>>> Randy Evans wrote: >>>> >>>>> I have created a shared link to my working folder with the details >>>>> of the HP-419A and Fluke 845A modifications. It is an in-process >>>>> design so I suspect there may be many questions. Just send your >>>>> questions and I will try to answer them. In the meantime, I will >>>>> try to find time to clean up the design a bit. >>>>> >>>>> Hopefully I created the link correctly. Let me know if you have >>>>> any issues with the link. >>>>> >>>>> https://www.dropbox.com/sh/7b3hsy0j0wq35hw/AABkmVptTwVu2EJFRqtSdAjya?dl=0 >>>>> >>>>> >>>>> Randy Evans AE6YG >>>>>