volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

3458A reference boards on ebay

JP
John Phillips
Thu, Jan 29, 2015 6:09 AM

I think teletekjm23@gmail.com has some that were removed from working unit
to upgrade to the more stable board.

On Wed, Jan 28, 2015 at 7:07 PM, Randy Evans randyevans2688@gmail.com
wrote:

I bought two HP-3458A ref boards from John Daly (he no longer has any more)
and they worked fine but have intermittent noise bursts.  I suspect they
were rejects from HP/Agilent.  At some point I will probably replace the
LTZ1000As and reduce the internal temperature set point for better
stability.  Has anyone else had any issues with these units?

BTW, I packaged the units in an enclosure for a stand-alone reference and,
except for the occasional noise bursts, seem very stable according to my HP
3458A and fluke 732A/ESI RV722 KVD.

Randy

On Wed, Jan 28, 2015 at 9:40 AM, John Phillips john.phillips0@gmail.com
wrote:

There is a bottom cover under the board.

On Tue, Jan 27, 2015 at 9:01 PM, J. L. Trantham jltran@att.net wrote:

When these boards are installed in a 3458A, the LTZ1000 is covered by a
small, white, 'hat' that mounts with screws that pass through the two

holes

on opposite sides of the LTZ1000.  I've never removed the board to see

if

there is anything on the bottom of the board that would cover the

bottom

of

the LTZ1000.

Does anyone know what this 'hat' is constructed of or the
HP/Agilent/Keysight part number of this item is?  Is it available from
Keysight?

In the 3458A the reference board is under a cover that covers about

half

of the top of the meter but not 'insulated' beyond that, except for the
'hat' referred to above.

It would appear desirable to find what that 'hat' is and see if

something

could be fabricated to 'insulate' the LTZ1000 for a 'do it yourself'
standalone reference.

Would appreciate hearing anything about how the reference board is
'insulated' in the 3458A.

Thanks.

Joe

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Orin

Eman

Sent: Tuesday, January 27, 2015 11:45 AM
To: Discussion of precise voltage measurement
Subject: [Bulk] Re: [volt-nuts] 3458A reference boards on ebay

Dave,

I'll be putting it in an enclosure to use as a reference.  It will
probably be a smallish diecast aluminium box.  It's what I did with my
Geller SVR-T and it compared very well with my Fluke 731B.  I'll likely

use

Pomona 3770 binding posts for output (I used them to repair the 731B).
I'll also use the protection circuits from the Fluke 732 on the output

big reverse diode, gas discharge tube and TVS as I recall (I plan to

add

these protection circuits to the 731B).

I'll likely do a PCB for it, though the local equivalent of veroboard
would likely be OK; the socket spacing is 0.9" and the pin spacing

0.1".

My meters are also 6 1/2 digit (3456A and 34461A) so I'm not too

concerned

with thermal EMFs with one count being 10uV on the 10V range.

However, I

will try to avoid temperature gradients and keep the outputs close
together.  I'll use star wiring for the 18V and ground as in the

following:

http://www.maxmcarter.com/vref/

It is debatable whether the -15V supply is needed - I'll test with and
without.  I'd add a few tantalum capacitors to the above design - on

the

regulator outputs and on the 3320 resistor, along with protection

diodes

for the LM317.

There is also a HUGE thread about the LTZ1000 on EEVBlog - I'll be
reviewing that for ideas:

I actually have a couple of LTZ1000As here that I was going to DIY, but
never got around to getting the precision resisters .  I think there

is a

source in England mentioned in the EEVBlog thread.

Orin.

On Tue, Jan 27, 2015 at 2:24 AM, Dr. David Kirkby (Kirkby Microwave

Ltd)

<

On 27 January 2015 at 09:16, Orin Eman orin.eman@gmail.com wrote:

There is a seller letting a slow trickle of them out there... they
seem

to

be going for about $165.  The current batch of two is around $100,
but

will

probably be bid up.

I just received one that I won.  Seems to be working fine after a
quick breadboard lashup.  I'll be making an enclosure for it next.

A little expensive IMO, but given the trouble getting the precision
resistors to DIY with the LTZ1000A, probably worth it.

Orin

Orin,

What do you intend doing with it?

I was thinking about getting one of those, with a view to putting it
in a box with a couple of terminals to have something to compare with
my 6.5 digit 3457A.  But what put me off is a lack of knowledge in
knowing how to convert a 3458A reference board into a boxed unit with
a known output voltage at the terminals. How would I avoid / control
thermal EMFs?

I'd be interested to hear what your plans are for it, and how you
intend tackling those issues.

Do you know what the difference in the reference is between a

standard

3458A (8 ppm) and the high stability option 002 (4 ppm) model? I'm
guessing the chips for the option 002 might be the top performing
ones. I wonder if there's any way to tell from your board if it came
from a standard 3458A or a 3458A with option 002.

Dave


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--

John Phillips


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--

John Phillips

I think teletekjm23@gmail.com has some that were removed from working unit to upgrade to the more stable board. On Wed, Jan 28, 2015 at 7:07 PM, Randy Evans <randyevans2688@gmail.com> wrote: > I bought two HP-3458A ref boards from John Daly (he no longer has any more) > and they worked fine but have intermittent noise bursts. I suspect they > were rejects from HP/Agilent. At some point I will probably replace the > LTZ1000As and reduce the internal temperature set point for better > stability. Has anyone else had any issues with these units? > > BTW, I packaged the units in an enclosure for a stand-alone reference and, > except for the occasional noise bursts, seem very stable according to my HP > 3458A and fluke 732A/ESI RV722 KVD. > > Randy > > On Wed, Jan 28, 2015 at 9:40 AM, John Phillips <john.phillips0@gmail.com> > wrote: > > > There is a bottom cover under the board. > > > > On Tue, Jan 27, 2015 at 9:01 PM, J. L. Trantham <jltran@att.net> wrote: > > > > > When these boards are installed in a 3458A, the LTZ1000 is covered by a > > > small, white, 'hat' that mounts with screws that pass through the two > > holes > > > on opposite sides of the LTZ1000. I've never removed the board to see > if > > > there is anything on the bottom of the board that would cover the > bottom > > of > > > the LTZ1000. > > > > > > Does anyone know what this 'hat' is constructed of or the > > > HP/Agilent/Keysight part number of this item is? Is it available from > > > Keysight? > > > > > > In the 3458A the reference board is under a cover that covers about > half > > > of the top of the meter but not 'insulated' beyond that, except for the > > > 'hat' referred to above. > > > > > > It would appear desirable to find what that 'hat' is and see if > something > > > could be fabricated to 'insulate' the LTZ1000 for a 'do it yourself' > > > standalone reference. > > > > > > Would appreciate hearing anything about how the reference board is > > > 'insulated' in the 3458A. > > > > > > Thanks. > > > > > > Joe > > > > > > > > > -----Original Message----- > > > From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Orin > > Eman > > > Sent: Tuesday, January 27, 2015 11:45 AM > > > To: Discussion of precise voltage measurement > > > Subject: [Bulk] Re: [volt-nuts] 3458A reference boards on ebay > > > > > > Dave, > > > > > > I'll be putting it in an enclosure to use as a reference. It will > > > probably be a smallish diecast aluminium box. It's what I did with my > > > Geller SVR-T and it compared very well with my Fluke 731B. I'll likely > > use > > > Pomona 3770 binding posts for output (I used them to repair the 731B). > > > I'll also use the protection circuits from the Fluke 732 on the output > - > > > big reverse diode, gas discharge tube and TVS as I recall (I plan to > add > > > these protection circuits to the 731B). > > > > > > I'll likely do a PCB for it, though the local equivalent of veroboard > > > would likely be OK; the socket spacing is 0.9" and the pin spacing > 0.1". > > > My meters are also 6 1/2 digit (3456A and 34461A) so I'm not too > > concerned > > > with thermal EMFs with one count being 10uV on the 10V range. > However, I > > > will try to avoid temperature gradients and keep the outputs close > > > together. I'll use star wiring for the 18V and ground as in the > > following: > > > > > > http://www.maxmcarter.com/vref/ > > > > > > It is debatable whether the -15V supply is needed - I'll test with and > > > without. I'd add a few tantalum capacitors to the above design - on > the > > > regulator outputs and on the 3320 resistor, along with protection > diodes > > > for the LM317. > > > > > > There is also a HUGE thread about the LTZ1000 on EEVBlog - I'll be > > > reviewing that for ideas: > > > > > > > http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/ > > > > > > I actually have a couple of LTZ1000As here that I was going to DIY, but > > > never got around to getting the precision resisters . I think there > is a > > > source in England mentioned in the EEVBlog thread. > > > > > > Orin. > > > > > > On Tue, Jan 27, 2015 at 2:24 AM, Dr. David Kirkby (Kirkby Microwave > Ltd) > > < > > > drkirkby@kirkbymicrowave.co.uk> wrote: > > > > > > > On 27 January 2015 at 09:16, Orin Eman <orin.eman@gmail.com> wrote: > > > > > There is a seller letting a slow trickle of them out there... they > > > > > seem > > > > to > > > > > be going for about $165. The current batch of two is around $100, > > > > > but > > > > will > > > > > probably be bid up. > > > > > > > > > > I just received one that I won. Seems to be working fine after a > > > > > quick breadboard lashup. I'll be making an enclosure for it next. > > > > > > > > > > A little expensive IMO, but given the trouble getting the precision > > > > > resistors to DIY with the LTZ1000A, probably worth it. > > > > > > > > > > Orin > > > > > > > > Orin, > > > > > > > > What do you intend doing with it? > > > > > > > > I was thinking about getting one of those, with a view to putting it > > > > in a box with a couple of terminals to have something to compare with > > > > my 6.5 digit 3457A. But what put me off is a lack of knowledge in > > > > knowing how to convert a 3458A reference board into a boxed unit with > > > > a known output voltage at the terminals. How would I avoid / control > > > > thermal EMFs? > > > > > > > > I'd be interested to hear what your plans are for it, and how you > > > > intend tackling those issues. > > > > > > > > Do you know what the difference in the reference is between a > standard > > > > 3458A (8 ppm) and the high stability option 002 (4 ppm) model? I'm > > > > guessing the chips for the option 002 might be the top performing > > > > ones. I wonder if there's any way to tell from your board if it came > > > > from a standard 3458A or a 3458A with option 002. > > > > > > > > Dave > > > > _______________________________________________ > > > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > > > and follow the instructions there. > > > > > > > _______________________________________________ > > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > > and follow the instructions there. > > > > > > _______________________________________________ > > > volt-nuts mailing list -- volt-nuts@febo.com > > > To unsubscribe, go to > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > > and follow the instructions there. > > > > > > > > > > > -- > > > > *John Phillips* > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- *John Phillips*
HD
Hendrik Dietrich
Thu, Jan 29, 2015 6:28 AM

Am 29.01.2015 04:07, schrieb Randy Evans:

I bought two HP-3458A ref boards from John Daly (he no longer has any more)
and they worked fine but have intermittent noise bursts.  I suspect they
were rejects from HP/Agilent.  At some point I will probably replace the
LTZ1000As and reduce the internal temperature set point for better
stability.  Has anyone else had any issues with these units?

BTW, I packaged the units in an enclosure for a stand-alone reference and,
except for the occasional noise bursts, seem very stable according to my HP
3458A and fluke 732A/ESI RV722 KVD.

Randy

Hi Randy,

Noise bursts, especially if they get back on track again, sound like
electromagnetic interference from your short description.
If applicable, turn of the CFL or LED lighting, replace the switch-mode
power supply with a linear one or even better a battery,, try a known
non-reject in place of the ebay reference module, spread some capacitors
on/around the module and/or invest in a metal box.

Hendrik

Am 29.01.2015 04:07, schrieb Randy Evans: > I bought two HP-3458A ref boards from John Daly (he no longer has any more) > and they worked fine but have intermittent noise bursts. I suspect they > were rejects from HP/Agilent. At some point I will probably replace the > LTZ1000As and reduce the internal temperature set point for better > stability. Has anyone else had any issues with these units? > > BTW, I packaged the units in an enclosure for a stand-alone reference and, > except for the occasional noise bursts, seem very stable according to my HP > 3458A and fluke 732A/ESI RV722 KVD. > > Randy > Hi Randy, Noise bursts, especially if they get back on track again, sound like electromagnetic interference from your short description. If applicable, turn of the CFL or LED lighting, replace the switch-mode power supply with a linear one or even better a battery,, try a known non-reject in place of the ebay reference module, spread some capacitors on/around the module and/or invest in a metal box. Hendrik
RE
Randy Evans
Thu, Jan 29, 2015 5:22 PM

The reference module is in a metal box and uses an external linear PS with
internal separate regulation.  I don't think its external EMI, but anything
is possible.

Randy

On Wed, Jan 28, 2015 at 10:28 PM, Hendrik Dietrich don_hendi@gmx.de wrote:

Am 29.01.2015 04:07, schrieb Randy Evans:

I bought two HP-3458A ref boards from John Daly (he no longer has any
more)
and they worked fine but have intermittent noise bursts.  I suspect they
were rejects from HP/Agilent.  At some point I will probably replace the
LTZ1000As and reduce the internal temperature set point for better
stability.  Has anyone else had any issues with these units?

BTW, I packaged the units in an enclosure for a stand-alone reference and,
except for the occasional noise bursts, seem very stable according to my
HP
3458A and fluke 732A/ESI RV722 KVD.

Randy

Hi Randy,

Noise bursts, especially if they get back on track again, sound like
electromagnetic interference from your short description.
If applicable, turn of the CFL or LED lighting, replace the switch-mode
power supply with a linear one or even better a battery,, try a known
non-reject in place of the ebay reference module, spread some capacitors
on/around the module and/or invest in a metal box.

Hendrik


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.

The reference module is in a metal box and uses an external linear PS with internal separate regulation. I don't think its external EMI, but anything is possible. Randy On Wed, Jan 28, 2015 at 10:28 PM, Hendrik Dietrich <don_hendi@gmx.de> wrote: > > Am 29.01.2015 04:07, schrieb Randy Evans: > >> I bought two HP-3458A ref boards from John Daly (he no longer has any >> more) >> and they worked fine but have intermittent noise bursts. I suspect they >> were rejects from HP/Agilent. At some point I will probably replace the >> LTZ1000As and reduce the internal temperature set point for better >> stability. Has anyone else had any issues with these units? >> >> BTW, I packaged the units in an enclosure for a stand-alone reference and, >> except for the occasional noise bursts, seem very stable according to my >> HP >> 3458A and fluke 732A/ESI RV722 KVD. >> >> Randy >> >> > Hi Randy, > > Noise bursts, especially if they get back on track again, sound like > electromagnetic interference from your short description. > If applicable, turn of the CFL or LED lighting, replace the switch-mode > power supply with a linear one or even better a battery,, try a known > non-reject in place of the ebay reference module, spread some capacitors > on/around the module and/or invest in a metal box. > > Hendrik > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. >
HD
Hendrik Dietrich
Thu, Jan 29, 2015 8:36 PM

Well, sounds like you already prepared against some of my suggestions.
Still there are possibilities to troubleshoot it - i cannot neither see
that you checked everything I listed nor that my list is complete and
you had much bigger ideas.
The datasheet gives some figures telling how much some influences, like
R variations (and as a result I variations) get supressed on the output -
I would go hunting for nodes where the noise bursts are biggest if all
preventive actions failed. Up to a certain point, troubleshooting is fun.

Hendrik

Am 29.01.2015 18:22, schrieb Randy Evans:

The reference module is in a metal box and uses an external linear PS with
internal separate regulation.  I don't think its external EMI, but anything
is possible.

Randy

On Wed, Jan 28, 2015 at 10:28 PM, Hendrik Dietrich don_hendi@gmx.de wrote:

Am 29.01.2015 04:07, schrieb Randy Evans:

I bought two HP-3458A ref boards from John Daly (he no longer has any
more)
and they worked fine but have intermittent noise bursts.  I suspect they
were rejects from HP/Agilent.  At some point I will probably replace the
LTZ1000As and reduce the internal temperature set point for better
stability.  Has anyone else had any issues with these units?

BTW, I packaged the units in an enclosure for a stand-alone reference and,
except for the occasional noise bursts, seem very stable according to my
HP
3458A and fluke 732A/ESI RV722 KVD.

Randy

Hi Randy,

Noise bursts, especially if they get back on track again, sound like
electromagnetic interference from your short description.
If applicable, turn of the CFL or LED lighting, replace the switch-mode
power supply with a linear one or even better a battery, try a known
non-reject in place of the ebay reference module, spread some capacitors
on/around the module and/or invest in a metal box.

Hendrik


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Well, sounds like you already prepared against some of my suggestions. Still there are possibilities to troubleshoot it - i cannot neither see that you checked everything I listed nor that my list is complete and you had much bigger ideas. The datasheet gives some figures telling how much some influences, like R variations (and as a result I variations) get supressed on the output - I would go hunting for nodes where the noise bursts are biggest if all preventive actions failed. Up to a certain point, troubleshooting is fun. Hendrik Am 29.01.2015 18:22, schrieb Randy Evans: > The reference module is in a metal box and uses an external linear PS with > internal separate regulation. I don't think its external EMI, but anything > is possible. > > Randy > > On Wed, Jan 28, 2015 at 10:28 PM, Hendrik Dietrich <don_hendi@gmx.de> wrote: > >> Am 29.01.2015 04:07, schrieb Randy Evans: >> >>> I bought two HP-3458A ref boards from John Daly (he no longer has any >>> more) >>> and they worked fine but have intermittent noise bursts. I suspect they >>> were rejects from HP/Agilent. At some point I will probably replace the >>> LTZ1000As and reduce the internal temperature set point for better >>> stability. Has anyone else had any issues with these units? >>> >>> BTW, I packaged the units in an enclosure for a stand-alone reference and, >>> except for the occasional noise bursts, seem very stable according to my >>> HP >>> 3458A and fluke 732A/ESI RV722 KVD. >>> >>> Randy >>> >>> >> Hi Randy, >> >> Noise bursts, especially if they get back on track again, sound like >> electromagnetic interference from your short description. >> If applicable, turn of the CFL or LED lighting, replace the switch-mode >> power supply with a linear one or even better a battery, try a known >> non-reject in place of the ebay reference module, spread some capacitors >> on/around the module and/or invest in a metal box. >> >> Hendrik >> >> >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/volt-nuts >> and follow the instructions there. >> > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Thu, Jan 29, 2015 8:59 PM

On 27 Jan 2015 09:47, "Orin Eman" orin.eman@gmail.com wrote:

There is a seller letting a slow trickle of them out there... they seem to
be going for about $165.  The current batch of two is around $100, but

will

probably be bid up.

I am puzzled why there are so many reference boards that are supposed to
come from 3458As on the used market. It makes me wonder if they are
genuine.

One seller,  with the two high stability reference boards at $750, claims
that they come from working and recently calibrated 3458As. Why would
anyone wreck a working and recently calibrated 3458A?

It is not beyond the bounds of possibility that a 3458A might be worth more
in parts than as a complete unit. I know this is true for some HP items,
but I doubt that is the case with a 3458A.

Another seller,  who has them at $395 "or offer", has sold 17 and has 11
left.

Although expensive,  having two and monitoring the difference between them
would probably allow their stability to be measured with a DVM of modest
specification.

There are lots of counterfeit components around,  and I fear that it could
be some of these boards could fall into that category.  This makes the
building a board from new components more attractive.  In the short term it
is likely to be less stable than an old reference,  but at least one would
know the history of it. If the components were sourced from reliable
sources, it would get around the  potential counterfeit problem.

Dave.

On 27 Jan 2015 09:47, "Orin Eman" <orin.eman@gmail.com> wrote: > > There is a seller letting a slow trickle of them out there... they seem to > be going for about $165. The current batch of two is around $100, but will > probably be bid up. I am puzzled why there are so many reference boards that are supposed to come from 3458As on the used market. It makes me wonder if they are genuine. One seller, with the two high stability reference boards at $750, claims that they come from working and recently calibrated 3458As. Why would anyone wreck a working and recently calibrated 3458A? It is not beyond the bounds of possibility that a 3458A might be worth more in parts than as a complete unit. I know this is true for some HP items, but I doubt that is the case with a 3458A. Another seller, who has them at $395 "or offer", has sold 17 and has 11 left. Although expensive, having two and monitoring the difference between them would probably allow their stability to be measured with a DVM of modest specification. There are lots of counterfeit components around, and I fear that it could be some of these boards could fall into that category. This makes the building a board from new components more attractive. In the short term it is likely to be less stable than an old reference, but at least one would know the history of it. If the components were sourced from reliable sources, it would get around the potential counterfeit problem. Dave.
JP
John Phillips
Thu, Jan 29, 2015 10:54 PM

RE: "One seller,  with the two high stability reference boards at $750,
claims
that they come from working and recently calibrated 3458As. Why would
anyone wreck a working and recently calibrated 3458A?
​"​

Some contracts may require the more stable reference and the standard
reference is pulled to make room for the option 002.

I know I have done that at lest 30 times.

On Thu, Jan 29, 2015 at 12:59 PM, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkirkby@kirkbymicrowave.co.uk> wrote:

On 27 Jan 2015 09:47, "Orin Eman" orin.eman@gmail.com wrote:

There is a seller letting a slow trickle of them out there... they seem

to

be going for about $165.  The current batch of two is around $100, but

will

probably be bid up.

I am puzzled why there are so many reference boards that are supposed to
come from 3458As on the used market. It makes me wonder if they are
genuine.

One seller,  with the two high stability reference boards at $750, claims
that they come from working and recently calibrated 3458As. Why would
anyone wreck a working and recently calibrated 3458A?

It is not beyond the bounds of possibility that a 3458A might be worth more
in parts than as a complete unit. I know this is true for some HP items,
but I doubt that is the case with a 3458A.

Another seller,  who has them at $395 "or offer", has sold 17 and has 11
left.

Although expensive,  having two and monitoring the difference between them
would probably allow their stability to be measured with a DVM of modest
specification.

There are lots of counterfeit components around,  and I fear that it could
be some of these boards could fall into that category.  This makes the
building a board from new components more attractive.  In the short term it
is likely to be less stable than an old reference,  but at least one would
know the history of it. If the components were sourced from reliable
sources, it would get around the  potential counterfeit problem.

Dave.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--

John Phillips

RE: "One seller, with the two high stability reference boards at $750, claims that they come from working and recently calibrated 3458As. Why would anyone wreck a working and recently calibrated 3458A? ​"​ Some contracts may require the more stable reference and the standard reference is pulled to make room for the option 002. I know I have done that at lest 30 times. On Thu, Jan 29, 2015 at 12:59 PM, Dr. David Kirkby (Kirkby Microwave Ltd) < drkirkby@kirkbymicrowave.co.uk> wrote: > On 27 Jan 2015 09:47, "Orin Eman" <orin.eman@gmail.com> wrote: > > > > There is a seller letting a slow trickle of them out there... they seem > to > > be going for about $165. The current batch of two is around $100, but > will > > probably be bid up. > > I am puzzled why there are so many reference boards that are supposed to > come from 3458As on the used market. It makes me wonder if they are > genuine. > > One seller, with the two high stability reference boards at $750, claims > that they come from working and recently calibrated 3458As. Why would > anyone wreck a working and recently calibrated 3458A? > > It is not beyond the bounds of possibility that a 3458A might be worth more > in parts than as a complete unit. I know this is true for some HP items, > but I doubt that is the case with a 3458A. > > Another seller, who has them at $395 "or offer", has sold 17 and has 11 > left. > > Although expensive, having two and monitoring the difference between them > would probably allow their stability to be measured with a DVM of modest > specification. > > There are lots of counterfeit components around, and I fear that it could > be some of these boards could fall into that category. This makes the > building a board from new components more attractive. In the short term it > is likely to be less stable than an old reference, but at least one would > know the history of it. If the components were sourced from reliable > sources, it would get around the potential counterfeit problem. > > Dave. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- *John Phillips*
W
Will
Fri, Jan 30, 2015 12:25 PM

The boards are factory rejects.

I bought five of them several years ago when the price was around $40.

One of them had the heater transistor flipped. Worked fine after
reversing the transistor. One had misaligned connector, which would
have been easy to repair but the HP policy seems to be reject rather
than repair small boards. One had corner of the Vishay resistor
cracked.

Two of them were cosmetically perfect but had much higher noise than others.

Be prepared that there is always something wrong. Could be a good idea
to prefer those with some obvious cosmetic problem...

I am puzzled why there are so many reference boards that are supposed to
come from 3458As on the used market. It makes me wonder if they are
genuine.

One seller,  with the two high stability reference boards at $750, claims
that they come from working and recently calibrated 3458As. Why would
anyone wreck a working and recently calibrated 3458A?

It is not beyond the bounds of possibility that a 3458A might be worth more
in parts than as a complete unit. I know this is true for some HP items,
but I doubt that is the case with a 3458A.

Another seller,  who has them at $395 "or offer", has sold 17 and has 11
left.

Although expensive,  having two and monitoring the difference between them
would probably allow their stability to be measured with a DVM of modest
specification.

There are lots of counterfeit components around,  and I fear that it could
be some of these boards could fall into that category.  This makes the
building a board from new components more attractive.  In the short term it
is likely to be less stable than an old reference,  but at least one would
know the history of it. If the components were sourced from reliable
sources, it would get around the  potential counterfeit problem.

Dave.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The boards are factory rejects. I bought five of them several years ago when the price was around $40. One of them had the heater transistor flipped. Worked fine after reversing the transistor. One had misaligned connector, which would have been easy to repair but the HP policy seems to be reject rather than repair small boards. One had corner of the Vishay resistor cracked. Two of them were cosmetically perfect but had much higher noise than others. Be prepared that there is always something wrong. Could be a good idea to prefer those with some obvious cosmetic problem... > I am puzzled why there are so many reference boards that are supposed to > come from 3458As on the used market. It makes me wonder if they are > genuine. > > One seller, with the two high stability reference boards at $750, claims > that they come from working and recently calibrated 3458As. Why would > anyone wreck a working and recently calibrated 3458A? > > It is not beyond the bounds of possibility that a 3458A might be worth more > in parts than as a complete unit. I know this is true for some HP items, > but I doubt that is the case with a 3458A. > > Another seller, who has them at $395 "or offer", has sold 17 and has 11 > left. > > Although expensive, having two and monitoring the difference between them > would probably allow their stability to be measured with a DVM of modest > specification. > > There are lots of counterfeit components around, and I fear that it could > be some of these boards could fall into that category. This makes the > building a board from new components more attractive. In the short term it > is likely to be less stable than an old reference, but at least one would > know the history of it. If the components were sourced from reliable > sources, it would get around the potential counterfeit problem. > > Dave. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Fri, Jan 30, 2015 1:24 PM

On 30 January 2015 at 12:25, Will willvolts@gmail.com wrote:

The boards are factory rejects.

I doubt all boards on eBay are in this category. Some are probably
quite genuine, some may be rejects, and it would never surprise me if
some are counterfeit. The problem is, I have no idea what ones are
what.

I don't know what the problems sourcing the parts for making a
reference would be, but if a number of people wanted to make one from
parts, perhaps someone could produce a PCB and people contribute
towards the cost of it. Maybe leave people to get their own reference
chip, as there are different grades of that. But things like the
precision resistors may be easier to buy in a quantity of 10 rather
than 1.

I would certainly not mind buying one if a PCB was available. Even if
it works out costing a bit more than a 3458A reference from eBay, I'd
feel a bit happier.

Dave

On 30 January 2015 at 12:25, Will <willvolts@gmail.com> wrote: > The boards are factory rejects. I doubt all boards on eBay are in this category. Some are probably quite genuine, some may be rejects, and it would never surprise me if some are counterfeit. The problem is, I have no idea what ones are what. I don't know what the problems sourcing the parts for making a reference would be, but if a number of people wanted to make one from parts, perhaps someone could produce a PCB and people contribute towards the cost of it. Maybe leave people to get their own reference chip, as there are different grades of that. But things like the precision resistors may be easier to buy in a quantity of 10 rather than 1. I would certainly not mind buying one if a PCB was available. Even if it works out costing a bit more than a 3458A reference from eBay, I'd feel a bit happier. Dave
TM
Todd Micallef
Fri, Jan 30, 2015 2:00 PM

There is currently a thread on EEVBlog where one has already been built. I
believe it can be purchased once the final layout is completed.

http://www.eevblog.com/forum/projects/project-kx-diy-calibrator-reference-sourcemeter/

Todd

On Fri, Jan 30, 2015 at 8:24 AM, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkirkby@kirkbymicrowave.co.uk> wrote:

On 30 January 2015 at 12:25, Will willvolts@gmail.com wrote:

The boards are factory rejects.

I doubt all boards on eBay are in this category. Some are probably
quite genuine, some may be rejects, and it would never surprise me if
some are counterfeit. The problem is, I have no idea what ones are
what.

I don't know what the problems sourcing the parts for making a
reference would be, but if a number of people wanted to make one from
parts, perhaps someone could produce a PCB and people contribute
towards the cost of it. Maybe leave people to get their own reference
chip, as there are different grades of that. But things like the
precision resistors may be easier to buy in a quantity of 10 rather
than 1.

I would certainly not mind buying one if a PCB was available. Even if
it works out costing a bit more than a 3458A reference from eBay, I'd
feel a bit happier.

Dave


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

There is currently a thread on EEVBlog where one has already been built. I believe it can be purchased once the final layout is completed. http://www.eevblog.com/forum/projects/project-kx-diy-calibrator-reference-sourcemeter/ Todd On Fri, Jan 30, 2015 at 8:24 AM, Dr. David Kirkby (Kirkby Microwave Ltd) < drkirkby@kirkbymicrowave.co.uk> wrote: > On 30 January 2015 at 12:25, Will <willvolts@gmail.com> wrote: > > The boards are factory rejects. > > I doubt all boards on eBay are in this category. Some are probably > quite genuine, some may be rejects, and it would never surprise me if > some are counterfeit. The problem is, I have no idea what ones are > what. > > I don't know what the problems sourcing the parts for making a > reference would be, but if a number of people wanted to make one from > parts, perhaps someone could produce a PCB and people contribute > towards the cost of it. Maybe leave people to get their own reference > chip, as there are different grades of that. But things like the > precision resistors may be easier to buy in a quantity of 10 rather > than 1. > > I would certainly not mind buying one if a PCB was available. Even if > it works out costing a bit more than a 3458A reference from eBay, I'd > feel a bit happier. > > Dave > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
DM
Dave M
Fri, Jan 30, 2015 3:33 PM

I'd go for a deal like that as well.  I have a small number of LM399s that
I'd like to put to use, as well as a couple LTZ1000A references.  A project
using either type of reference would be very exciting for me.
I have a few old Fluke differential meters that have lots of nice low
tempco, high quality resistors that might finally find a useful project such
as this.

Count me in,
Dave M.

Dr. David Kirkby (Kirkby Microwave Ltd) wrote:

On 30 January 2015 at 12:25, Will willvolts@gmail.com wrote:

The boards are factory rejects.

I doubt all boards on eBay are in this category. Some are probably
quite genuine, some may be rejects, and it would never surprise me if
some are counterfeit. The problem is, I have no idea what ones are
what.

I don't know what the problems sourcing the parts for making a
reference would be, but if a number of people wanted to make one from
parts, perhaps someone could produce a PCB and people contribute
towards the cost of it. Maybe leave people to get their own reference
chip, as there are different grades of that. But things like the
precision resistors may be easier to buy in a quantity of 10 rather
than 1.

I would certainly not mind buying one if a PCB was available. Even if
it works out costing a bit more than a 3458A reference from eBay, I'd
feel a bit happier.

Dave

I'd go for a deal like that as well. I have a small number of LM399s that I'd like to put to use, as well as a couple LTZ1000A references. A project using either type of reference would be very exciting for me. I have a few old Fluke differential meters that have lots of nice low tempco, high quality resistors that might finally find a useful project such as this. Count me in, Dave M. Dr. David Kirkby (Kirkby Microwave Ltd) wrote: > On 30 January 2015 at 12:25, Will <willvolts@gmail.com> wrote: >> The boards are factory rejects. > > I doubt all boards on eBay are in this category. Some are probably > quite genuine, some may be rejects, and it would never surprise me if > some are counterfeit. The problem is, I have no idea what ones are > what. > > I don't know what the problems sourcing the parts for making a > reference would be, but if a number of people wanted to make one from > parts, perhaps someone could produce a PCB and people contribute > towards the cost of it. Maybe leave people to get their own reference > chip, as there are different grades of that. But things like the > precision resistors may be easier to buy in a quantity of 10 rather > than 1. > > I would certainly not mind buying one if a PCB was available. Even if > it works out costing a bit more than a 3458A reference from eBay, I'd > feel a bit happier. > > Dave