Hello all,
I agree that the M version was made for military customers only, being a US
Navy contract in about 1990. Aside from the thermal converter Wideband V
input, key differences I am aware of between 4920 and 4920M are that the M
version does not have mV ranges below 0.3V, and the M version has Filter
options of only 100Hz and 10Hz (not additional 40Hz). Programming of Filter
uses different syntax too.
After a question from another member I have scanned my paper copy of the D
4920M Users manual, but its a bit of a mess. I will take it to work
tomorrow and re-do it again properly, then upload it to KO4BB and BAMA.
Hopefully the Calibration and Service manual will follow in short order.
I can confirm that calibration of values 0.3V to 1000V use a thermal
converter. As well, calibration of 100mV (0.3V range) will require a
micropot and DC voltmeter.
--- I have an alternative version of this email sitting in the Moderators
in-tray which has attached a PDF of a paper presented at MSC 1990
conference (202kB file size). If returned to me I will find another site to
post the paper - it makes good reading. If interested please email me and
I can send it out individually. Otherwise, the whole MSC 1990 conference
proceedings, including the Datron 4920M paper, can be found here ->
msc-conf.com/wp-content/uploads/proceedings/MSC1990.pdf
More to follow,
regards,
ben
From: "Stephen Grady" grady.steve@gmail.com
Sent: Tuesday, September 02, 2014 7:54 PM
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1
day left
The Datron 4920M was a special version of the 4920 made for the military,
the major difference is the M version has a 50 Ohm (N-female) wideband
input and only the N-female input for the low frequency input. Whereas the
4920 has a Binding Post Input as well as the N-female input. I am not aware
of the specs for the wideband input but the low frequency input will be the
same for both models.
I intend to reply in some detail soon on the calibration of these
instruments (and this will also apply to the Fluke 5790) and the principals
of AC calibration below 100mV as Charles has raised the issue of using
attenuators to do lower voltage measurements.
Kind Regards,
Stephen Grady
Sydney, Australia
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of
acbern@gmx.de
Sent: Tuesday, 2 September 2014 5:54 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1
day left
folks,
does anybody know the difference in specs between a 4920M and the standard
4920. No data on the web.
thanks
Gesendet: Sonntag, 31. August 2014 um 23:41 Uhr
Von: "Charles Black"
An: "Discussion of precise voltage measurement"
Betreff: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay
Hi Stephen,
Thanks for the info on the 4920M. It might be that I should pass on
this but it is nice to know that it is probably working. If I don't
have the necessary items needed for calibration it might still be a
good source uncalibrated.
I was hoping that the 4920 had better minimum voltage out steps than
my present Rigol 4062 for 3458a AC calibration. I have two EL 1400
0.25V voltage thermal converters that I was trying to use in a test
setup but they had such a low voltage range that I was forced to use
my ATV-60 attenuator. It all kind of worked but a 0.1 db minimum step
is a little course for easy use. The Rigol also suffers from this malady.
Charlie
On 8/31/2014 1:06 PM, Stephen Grady wrote:
Charlie,
The "Error Ur" is an under-range error; you have to apply an input
above 10% of the range before it will display a reading. The 4920's are a
very nice instrument. Their only problem is that they are all so old that
they are reaching the stage where some components are drifting excessively
or failing. I have come across 4920's that are drifting a little (more than
there spec) between cals (and yes they need to be cal'd annually), another
4920 I came across had an intermittent failure in one of its power
supplies.
Kind Regards,
Stephen Grady
Sydney Australia
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of
Charles Black
Sent: Monday, 1 September 2014 3:19 AM
To: ben@veritechmeasurements.com.au; Discussion of precise voltage
measurement
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on
ebay - 1 day left
Hi Ben,
I am going to check this out today to find if it might be helpfull
around here for my AC calibration. It has an error message: "Error Ur"
so some fixing is going to be needed. On the back it says it's got
Option 80, whatever that is.
Charlie
On 8/31/2014 5:37 AM, ben wrote:
Hello all,
Just a heads up. I noted there is a lonely Datron / Wavetek 4920M
AC voltmeter for sale on ebay, starting at US$650 (ending in 1 day
!). I have little idea of its true working state or not (picture shows
it powered on).
If fully working it would be a good buy for an AC voltmeter that
is, I reckon, better than an HP 3458A. Only drawback with these
models is the voltage input shell is always earthed, not floating.
I have two of these 4920M's already, not really tempted by a third.
I have a paper copies of operator, calibration, and service manual if
anyone interested.
regards, ben.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
This email is free from viruses and malware because avast! Antivirus
protection is active.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
This email is free from viruses and malware because avast! Antivirus
protection is active.
http://www.avast.com
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I uploaded the user manual for the 4920 model some time ago. Of note, the 4920 was offered with option 10. This added four additional millivolt ranges. I have this meter with the option 10. I think it may be a better alternative to the 4920m if you don't need the wideband option. Some of the 4920 had an external wideband input jack, but I have not seen this adapter for sale. It appears these were designed to compete with the Fluke 540B due to its AC/DC transfer feature.
My meter has a limited calibration due to a couple ranges not meeting spec. Unfortunately, only the 4920M service manual is available. If anyone has the service manual for this, it would be appreciated.
Todd
Sent from my iPad
On Sep 2, 2014, at 10:17, "ben" ben@veritechmeasurements.com.au wrote:
From: "ben" ben@veritechmeasurements.com.au
Sent: Tuesday, September 02, 2014 11:25 PM
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1
day left
Hello all,
I agree that the M version was made for military customers only, being a US
Navy contract in about 1990. Aside from the thermal converter Wideband V
input, key differences I am aware of between 4920 and 4920M are that the M
version does not have mV ranges below 0.3V, and the M version has Filter
options of only 100Hz and 10Hz (not additional 40Hz). Programming of Filter
uses different syntax too.
After a question from another member I have scanned my paper copy of the D
4920M Users manual, but its a bit of a mess. I will take it to work
tomorrow and re-do it again properly, then upload it to KO4BB and BAMA.
Hopefully the Calibration and Service manual will follow in short order.
I can confirm that calibration of values 0.3V to 1000V use a thermal
converter. As well, calibration of 100mV (0.3V range) will require a
micropot and DC voltmeter.
--- I have an alternative version of this email sitting in the Moderators
in-tray which has attached a PDF of a paper presented at MSC 1990
conference (202kB file size). If returned to me I will find another site to
post the paper - it makes good reading. If interested please email me and
I can send it out individually. Otherwise, the whole MSC 1990 conference
proceedings, including the Datron 4920M paper, can be found here ->
msc-conf.com/wp-content/uploads/proceedings/MSC1990.pdf
More to follow,
regards,
ben
From: "Stephen Grady" grady.steve@gmail.com
Sent: Tuesday, September 02, 2014 7:54 PM
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1
day left
The Datron 4920M was a special version of the 4920 made for the military,
the major difference is the M version has a 50 Ohm (N-female) wideband
input and only the N-female input for the low frequency input. Whereas the
4920 has a Binding Post Input as well as the N-female input. I am not aware
of the specs for the wideband input but the low frequency input will be the
same for both models.
I intend to reply in some detail soon on the calibration of these
instruments (and this will also apply to the Fluke 5790) and the principals
of AC calibration below 100mV as Charles has raised the issue of using
attenuators to do lower voltage measurements.
Kind Regards,
Stephen Grady
Sydney, Australia
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of
acbern@gmx.de
Sent: Tuesday, 2 September 2014 5:54 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1
day left
folks,
does anybody know the difference in specs between a 4920M and the standard
4920. No data on the web.
thanks
Gesendet: Sonntag, 31. August 2014 um 23:41 Uhr
Von: "Charles Black"
An: "Discussion of precise voltage measurement"
Betreff: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay
Hi Stephen,
Thanks for the info on the 4920M. It might be that I should pass on
this but it is nice to know that it is probably working. If I don't
have the necessary items needed for calibration it might still be a
good source uncalibrated.
I was hoping that the 4920 had better minimum voltage out steps than
my present Rigol 4062 for 3458a AC calibration. I have two EL 1400
0.25V voltage thermal converters that I was trying to use in a test
setup but they had such a low voltage range that I was forced to use
my ATV-60 attenuator. It all kind of worked but a 0.1 db minimum step
is a little course for easy use. The Rigol also suffers from this malady.
Charlie
On 8/31/2014 1:06 PM, Stephen Grady wrote:
Charlie,
The "Error Ur" is an under-range error; you have to apply an input
above 10% of the range before it will display a reading. The 4920's are a
very nice instrument. Their only problem is that they are all so old that
they are reaching the stage where some components are drifting excessively
or failing. I have come across 4920's that are drifting a little (more than
there spec) between cals (and yes they need to be cal'd annually), another
4920 I came across had an intermittent failure in one of its power
supplies.
Kind Regards,
Stephen Grady
Sydney Australia
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of
Charles Black
Sent: Monday, 1 September 2014 3:19 AM
To: ben@veritechmeasurements.com.au; Discussion of precise voltage
measurement
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on
ebay - 1 day left
Hi Ben,
I am going to check this out today to find if it might be helpfull
around here for my AC calibration. It has an error message: "Error Ur"
so some fixing is going to be needed. On the back it says it's got
Option 80, whatever that is.
Charlie
On 8/31/2014 5:37 AM, ben wrote:
Hello all,
Just a heads up. I noted there is a lonely Datron / Wavetek 4920M
AC voltmeter for sale on ebay, starting at US$650 (ending in 1 day
!). I have little idea of its true working state or not (picture shows
it powered on).
If fully working it would be a good buy for an AC voltmeter that
is, I reckon, better than an HP 3458A. Only drawback with these
models is the voltage input shell is always earthed, not floating.
I have two of these 4920M's already, not really tempted by a third.
I have a paper copies of operator, calibration, and service manual if
anyone interested.
regards, ben.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
This email is free from viruses and malware because avast! Antivirus
protection is active.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
This email is free from viruses and malware because avast! Antivirus
protection is active.
http://www.avast.com
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.