I am working on a Datron 1281 meter.
DC voltage calibration at 10v functions ok, but the Internal source
calibration failed.
This Meter has the following errors after Internal Source Calibration.
Error codes:
2202 1KV Range zero magnitude test (P028).
2192 100V Range zero magnitude test (P021).
2162 10V Range zero magnitude test (P006).
What would cause these errors?
M202 (DG211) switchs, pins 6-8 and 1-3.
Friday, March 21, 2014, 2:45:46 PM, you wrote:
sun> I am working on a Datron 1281 meter.
sun> DC voltage calibration at 10v functions ok, but the Internal source
sun> calibration failed.
sun> This Meter has the following errors after Internal Source Calibration.
sun> Error codes:
sun> 2202 1KV Range zero magnitude test (P028).
sun> 2192 100V Range zero magnitude test (P021).
sun> 2162 10V Range zero magnitude test (P006).
sun> What would cause these errors?
Михаил <timka2k@...> writes:
M202 (DG211) switchs, pins 6-8 and 1-3.
Friday, March 21, 2014, 2:45:46 PM, you wrote:
sun> I am working on a Datron 1281 meter.
sun> DC voltage calibration at 10v functions ok, but the Internal source
sun> calibration failed.
sun> This Meter has the following errors after Internal Source Calibration.
sun> Error codes:
sun> 2202 1KV Range zero magnitude test (P028).
sun> 2192 100V Range zero magnitude test (P021).
sun> 2162 10V Range zero magnitude test (P006).
sun> What would cause these errors?
I have the same issue and prior to seeing this blog i checked the m202 dg11
switchs and did not find an issue. What i did see was noise on the guard
around that line which goes to the non-inverting side of the dc amp, where
that noise is coming from possibly the bootstrap i am not 100% or i would have
fixed it. But it seems like this is a common issue this is the second post i
have found on this issue which makes me believe it not a cold solder issue.
Ive checked most caps in power supply and dc board if you have any other
possibilities please let me know.