time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Re: [time-nuts] TIC resolution impact on GPSDO's performance

DB
Dr Bruce Griffiths
Tue, Dec 26, 2006 12:15 PM

Magnus Danielson wrote:

From: "Poul-Henning Kamp" phk@phk.freebsd.dk
Subject: Re: [time-nuts] TIC resolution impact on GPSDO's performance
Date: Tue, 26 Dec 2006 08:14:12 +0000
Message-ID: 36958.1167120852@critter.freebsd.dk

In message 4590BB56.5070809@xtra.co.nz, Dr Bruce Griffiths writes:

Poul-Henning Kamp wrote:

I think what you are trying to express is that the frequency from
the internal Xtal (at times) is in an overtone of the 1Hz PPS, which
gives rise to hanging bridges.

Poul-Henning

Surely you mean harmonic? An overtone is not necessarily a harmonic.

I meant overtone, becuase they are not in any harmonic relationship,
the XO is wandering around whereas the (ideal) PPS is, supposedly,
rock stable.

The only real way to describe it is to say that the XO and the PPS is
asynchronous to each other. Those who has freshen up on their greek for
technicians will know that this means "not the same clock" which is quite
accuratly what we have. ITU-T Rec. G.700 has a few interesting sections in it.

The use of words such as harmonic or overtone should not be used since they
is normally used to describe the various frequencies within one signal where
as words as synchronous and asynchronous is to be used mainly for pair of
signals and describes their relative timing. Naturally, for a pair of signals
to be synchronous does not require them to have the same frequency, just a
fixed ratio between those frequencies.

It does happends that a single signal is said to be synchronous, but the
wording which should have been used is isochronous (same clock). Then we have
the lovely words of plesiochronous and mesochronous. :)

Cheers,
Magnus


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Magnus

So the correct description of the cause of hanging bridges is something
like?

As the frequency of the clock which controls the timing of the receiver
PPS pulse output drifts and approaches pleisiosynchronism with the
"true" frequency of the PPS signal, the rate of change of the PPS timing
error due to quantisation decreases reaching zero should the clock
achieve synchronism with the "true" PPS frequency.
Thus hanging bridges occur when the clock frequency approaches
pleisiosynchronism with the "true" PPS signal frequency and then drifts
away from pleisiosynchronism with the PPS signal..

When the frequency of the clock controlling the timing of the receiver
PPS pulse output is synchronous with the PPS frequency, then neither
sawtooth corrections or hanging bridges occur, however there may be a
fixed offset of upto 1/2 a clock cycle between the "correct" and the
actual leading edge of the receiver PPS output pulse.

Please correct and simplify these statements as required.

The only problem that I can see is we still need to define how close to
synchronism the timing clock frequency and the "true" PPS signal
frequency system must be to be considered pleisiosynchronous.

Bruce

Magnus Danielson wrote: > From: "Poul-Henning Kamp" <phk@phk.freebsd.dk> > Subject: Re: [time-nuts] TIC resolution impact on GPSDO's performance > Date: Tue, 26 Dec 2006 08:14:12 +0000 > Message-ID: <36958.1167120852@critter.freebsd.dk> > > >> In message <4590BB56.5070809@xtra.co.nz>, Dr Bruce Griffiths writes: >> >>> Poul-Henning Kamp wrote: >>> >>> >>>> I think what you are trying to express is that the frequency from >>>> the internal Xtal (at times) is in an overtone of the 1Hz PPS, which >>>> gives rise to hanging bridges. >>>> >>>> >>>> >>> Poul-Henning >>> >>> Surely you mean harmonic? An overtone is not necessarily a harmonic. >>> >> I meant overtone, becuase they are not in any harmonic relationship, >> the XO is wandering around whereas the (ideal) PPS is, supposedly, >> rock stable. >> > > The only real way to describe it is to say that the XO and the PPS is > asynchronous to each other. Those who has freshen up on their greek for > technicians will know that this means "not the same clock" which is quite > accuratly what we have. ITU-T Rec. G.700 has a few interesting sections in it. > > The use of words such as harmonic or overtone should not be used since they > is normally used to describe the various frequencies within one signal where > as words as synchronous and asynchronous is to be used mainly for pair of > signals and describes their relative timing. Naturally, for a pair of signals > to be synchronous does not require them to have the same frequency, just a > fixed ratio between those frequencies. > > It does happends that a single signal is said to be synchronous, but the > wording which should have been used is isochronous (same clock). Then we have > the lovely words of plesiochronous and mesochronous. :) > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > Magnus So the correct description of the cause of hanging bridges is something like? As the frequency of the clock which controls the timing of the receiver PPS pulse output drifts and approaches pleisiosynchronism with the "true" frequency of the PPS signal, the rate of change of the PPS timing error due to quantisation decreases reaching zero should the clock achieve synchronism with the "true" PPS frequency. Thus hanging bridges occur when the clock frequency approaches pleisiosynchronism with the "true" PPS signal frequency and then drifts away from pleisiosynchronism with the PPS signal.. When the frequency of the clock controlling the timing of the receiver PPS pulse output is synchronous with the PPS frequency, then neither sawtooth corrections or hanging bridges occur, however there may be a fixed offset of upto 1/2 a clock cycle between the "correct" and the actual leading edge of the receiver PPS output pulse. Please correct and simplify these statements as required. The only problem that I can see is we still need to define how close to synchronism the timing clock frequency and the "true" PPS signal frequency system must be to be considered pleisiosynchronous. Bruce
PK
Poul-Henning Kamp
Tue, Dec 26, 2006 1:18 PM

In message 20061226.125102.1723212534.cfmd@bredband.net, Magnus Danielson wri
tes:

harmonic(2)
1a) a tone in a harmonic series
2) in physics, a component frequency of a harmonic motion that is an integral
multiple of the fundamental frequency

overtone
1b) Harmonic(2)

Overtone is a musical term and it is not the same as harmonic.

An harmonic frequency is N times the fundamental, where as an
overtone merely holds a rational relationship to the fundamental
(3/2, 4/3 etc.)

The reason you can tell the difference between a violin and a
clarinet is that they have different overtone spectra.

A somewhat arbitrary subset of an intruments overtones are called
"formants" because they are the principal "formers" of the instruments
sound.

synchronous

  1. happening, existing or arising at precisely the same time
    2a) going on or operating together at exactly the same rate
    b) recurring together
  2. involving or indicating synchronism
  1. in physics, having the same period or having the same period and phase

Close, but no cigar.

If they merely have the same period/frequency, they are "syntonous",
but if the phase is also coincident, they are synchronous.

asynchronous
not synchronous, proceeding at its own pace;

But just because two signals are asynchronous, doesn't mean they cannot
have the same frequency or phase for a shorter or longer period of time,
it just means that there is no mechanism that makes it so.

The "asynchronous" serial line is actually anisochronous BTW.

No, each transmitter and receiver are anisochronous, but the
communication between them, and by language extension the actual
cable, is asynchronous.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <20061226.125102.1723212534.cfmd@bredband.net>, Magnus Danielson wri tes: >harmonic(2) >1a) a tone in a harmonic series >2) in physics, a component frequency of a harmonic motion that is an integral > multiple of the fundamental frequency > >overtone >1b) Harmonic(2) Overtone is a musical term and it is not the same as harmonic. An harmonic frequency is N times the fundamental, where as an overtone merely holds a rational relationship to the fundamental (3/2, 4/3 etc.) The reason you can tell the difference between a violin and a clarinet is that they have different overtone spectra. A somewhat arbitrary subset of an intruments overtones are called "formants" because they are the principal "formers" of the instruments sound. >synchronous >1) happening, existing or arising at precisely the same time >2a) going on or operating together at exactly the same rate > b) recurring together >3) involving or indicating synchronism >4) in physics, having the same period or having the same period and phase Close, but no cigar. If they merely have the same period/frequency, they are "syntonous", but if the phase is also coincident, they are synchronous. >asynchronous >not synchronous, proceeding at its own pace; But just because two signals are asynchronous, doesn't mean they cannot have the same frequency or phase for a shorter or longer period of time, it just means that there is no mechanism that makes it so. >The "asynchronous" serial line is actually anisochronous BTW. No, each transmitter and receiver are anisochronous, but the communication between them, and by language extension the actual cable, is asynchronous. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
DJ
Didier Juges
Tue, Dec 26, 2006 3:40 PM

Dr Bruce Griffiths wrote:

Didier Juges wrote:

This discussion is fascinating, and as always it has prompted a number
of other questions for me.

I understand the sawtooth correction is provided to allow correction of
1 PPS timing errors when the processor clock is non-coherent with the
GPS signal (at least not intentionally), because the processor clock is
running at a finite frequency and provides granularity to the PPS
adjustment capability.

I wanted to know if this was available in my Trimble Thunderbolt GPSDO.

I have looked over the entire Thunderbolt manual and I have not seen a
mention about sawtooth correction, so I am not sure if it is available.

Then I realized the Thunderbolt is different from most GPSDO that have
been discussed, in that it uses the 10 MHz from the OCXO as the clock
for the GPS receiver, so in theory, the sawtooth correction should not
be necessary on this type of receiver since the processor clock and the
GPS signals are coherent (maybe I should not use that term...) when the
receiver is tracking.

Yet, the Thunderbolt spec for timing accuracy  on the 1 PPS output is
+/- 20 nS at 1 sigma, just like an ordinary GPSDO with stand alone
receiver and what would be about a 25 MHz clock (please confirm).

The Trimble software also displays values referred to as "Timing Output"
for the 1 PPS and the 10 MHz. The PPS value is currently hovering around
-1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x ppb. I am not
sure what that is based on. The manual simply refers to those as
"Estimate of UTC/GPS offset", but does not give any detail of how they
are computed or what they mean.

The Thunderbolt User's Manual says that the 1 PPS output is the OCXO's
10 MHz divided down. Yet, the block diagram shows the 1 PPS coming from
the CPU and support circuit, not directly coming from the 10 MHz. I
believe the 1 PPS is the 10 MHz divided down, except when the receiver
is doing jam-sync (after power up or recovery, when the GPS_PPS and
HW_PPS are far away from each other)

Going back to what Poul-Henning just said, I believe that for the
Thunderbolt, there actually may be three PPS signals:

GPS_PPS: what the receiver thinks the PPS should be (a theoretical signal)
HW_PPS: the best approximation of the GPS_PPS that the receiver can do
OCXO_PPS: the 10 MHz divided down

In jam-sync mode, the Thunderbolt forces the HW_PPS to within 100nS of
GPS_PPS (the closest it can get using software delays and programmable
dividers I guess, using a 10 MHz clock) without touching the OCXO, so in
that mode, HW_PPS and OCXO_PPS are obviously not the same. Once the two
are within 100 nS, the Thunderbolt switches to discipline mode and fine
tunes the OCXO to get the HW_PPS as close to GPS_PPS as possible (20nS 1
sigma by specification.) It would appear that the receiver should be
able to adjust its OCXO as close to GPS_PPS as the DAC will allow,
without the artificial limit of the processor clock period. So, there
should not be a need for sawtooth correction and there should be no
hanging bridges either.

Then, what are the "Timing Outputs"?

Does this make any sense?

Didier KO4BB


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Didier

Look at:
http://trl.trimble.com/docushare/dsweb/Get/Document-8425/

There is little evidence of sawtooth corrections in graph in the above
document, however the data were taken when SA was switched on.
To confuse matters further the document actually talks about pulse
position quantisation.

http://trl.trimble.com/docushare/dsweb/Get/Document-8424/
Shows differences between 2 Thunderbolts, again there is little evidence
of quantisation but data may have been smoothed.

Have you any reasonably stable crystal oscillator that can be used to
log the time interval between the Thunderbolt PPS output and the divided
down oscillator output?
Plotting these results should quickly show evidence of any sawtooth error.

As you say there shouldn't be any sawtooth error if the receiver timing
is derived from the 10MHz clock which is in turned locked to the PPS output.
However there are other sources of noise in a GPS timing receiver which
could, in an older receiver design like this, easily be as much as 20ns rms.

Bruce


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Bruce,

The Trimble app note you pointed to says the 1 PPS output is not
quantized, so I believe that means there should not be sawtooth error.

Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html

The last picture is the Thunderbolt PPS output against a free running HP
10811 divided by 256 with two 74F161.

The plot was obtained after removing steps, outliers (few, mostly due to
my acquisition software that puts some garbage in the output) and drift,
using Ulrich's Plotter program.

Since I was using the HP 5334, data is taken every 2 seconds.

How can I determine from this if sawtooth error is present?

Do I need to collect data differently?

Thanks

Didier

Note: my home brew GPIB data logging software records ambient
temperature at the same rate as the TI, which I believe would be a very
useful information. Unfortunately, I have not found yet how to display
both curves in a meaningful way using Plotter, so I have to strip the
temperature from the dataset before feeding it to Plotter. Also,
collecting data from two instruments causes my program to sometimes
insert garbage in the output file, which causes outliers and which I
need to fix before I put this software on my web page.

Dr Bruce Griffiths wrote: > Didier Juges wrote: > >> This discussion is fascinating, and as always it has prompted a number >> of other questions for me. >> >> I understand the sawtooth correction is provided to allow correction of >> 1 PPS timing errors when the processor clock is non-coherent with the >> GPS signal (at least not intentionally), because the processor clock is >> running at a finite frequency and provides granularity to the PPS >> adjustment capability. >> >> I wanted to know if this was available in my Trimble Thunderbolt GPSDO. >> >> I have looked over the entire Thunderbolt manual and I have not seen a >> mention about sawtooth correction, so I am not sure if it is available. >> >> Then I realized the Thunderbolt is different from most GPSDO that have >> been discussed, in that it uses the 10 MHz from the OCXO as the clock >> for the GPS receiver, so in theory, the sawtooth correction should not >> be necessary on this type of receiver since the processor clock and the >> GPS signals are coherent (maybe I should not use that term...) when the >> receiver is tracking. >> >> Yet, the Thunderbolt spec for timing accuracy on the 1 PPS output is >> +/- 20 nS at 1 sigma, just like an *ordinary* GPSDO with stand alone >> receiver and what would be about a 25 MHz clock (please confirm). >> >> The Trimble software also displays values referred to as "Timing Output" >> for the 1 PPS and the 10 MHz. The PPS value is currently hovering around >> -1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x ppb. I am not >> sure what that is based on. The manual simply refers to those as >> "Estimate of UTC/GPS offset", but does not give any detail of how they >> are computed or what they mean. >> >> The Thunderbolt User's Manual says that the 1 PPS output is the OCXO's >> 10 MHz divided down. Yet, the block diagram shows the 1 PPS coming from >> the CPU and support circuit, not directly coming from the 10 MHz. I >> believe the 1 PPS is the 10 MHz divided down, except when the receiver >> is doing jam-sync (after power up or recovery, when the GPS_PPS and >> HW_PPS are far away from each other) >> >> Going back to what Poul-Henning just said, I believe that for the >> Thunderbolt, there actually may be three PPS signals: >> >> GPS_PPS: what the receiver thinks the PPS should be (a theoretical signal) >> HW_PPS: the best approximation of the GPS_PPS that the receiver can do >> OCXO_PPS: the 10 MHz divided down >> >> In jam-sync mode, the Thunderbolt forces the HW_PPS to within 100nS of >> GPS_PPS (the closest it can get using software delays and programmable >> dividers I guess, using a 10 MHz clock) without touching the OCXO, so in >> that mode, HW_PPS and OCXO_PPS are obviously not the same. Once the two >> are within 100 nS, the Thunderbolt switches to discipline mode and fine >> tunes the OCXO to get the HW_PPS as close to GPS_PPS as possible (20nS 1 >> sigma by specification.) It would appear that the receiver should be >> able to adjust its OCXO as close to GPS_PPS as the DAC will allow, >> without the artificial limit of the processor clock period. So, there >> should not be a need for sawtooth correction and there should be no >> hanging bridges either. >> >> Then, what are the "Timing Outputs"? >> >> Does this make any sense? >> >> Didier KO4BB >> >> >> >> >> _______________________________________________ >> time-nuts mailing list >> time-nuts@febo.com >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> > Didier > > Look at: > http://trl.trimble.com/docushare/dsweb/Get/Document-8425/ > > There is little evidence of sawtooth corrections in graph in the above > document, however the data were taken when SA was switched on. > To confuse matters further the document actually talks about pulse > position quantisation. > > http://trl.trimble.com/docushare/dsweb/Get/Document-8424/ > Shows differences between 2 Thunderbolts, again there is little evidence > of quantisation but data may have been smoothed. > > Have you any reasonably stable crystal oscillator that can be used to > log the time interval between the Thunderbolt PPS output and the divided > down oscillator output? > Plotting these results should quickly show evidence of any sawtooth error. > > As you say there shouldn't be any sawtooth error if the receiver timing > is derived from the 10MHz clock which is in turned locked to the PPS output. > However there are other sources of noise in a GPS timing receiver which > could, in an older receiver design like this, easily be as much as 20ns rms. > > Bruce > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > Bruce, The Trimble app note you pointed to says the 1 PPS output is not quantized, so I believe that means there should not be sawtooth error. Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html The last picture is the Thunderbolt PPS output against a free running HP 10811 divided by 256 with two 74F161. The plot was obtained after removing steps, outliers (few, mostly due to my acquisition software that puts some garbage in the output) and drift, using Ulrich's Plotter program. Since I was using the HP 5334, data is taken every 2 seconds. How can I determine from this if sawtooth error is present? Do I need to collect data differently? Thanks Didier Note: my home brew GPIB data logging software records ambient temperature at the same rate as the TI, which I believe would be a very useful information. Unfortunately, I have not found yet how to display both curves in a meaningful way using Plotter, so I have to strip the temperature from the dataset before feeding it to Plotter. Also, collecting data from two instruments causes my program to sometimes insert garbage in the output file, which causes outliers and which I need to fix before I put this software on my web page.
DJ
Didier Juges
Tue, Dec 26, 2006 3:50 PM

Poul-Henning Kamp wrote:

In message 20061226.125102.1723212534.cfmd@bredband.net, Magnus Danielson wri
tes:

harmonic(2)
1a) a tone in a harmonic series
2) in physics, a component frequency of a harmonic motion that is an integral
multiple of the fundamental frequency

overtone
1b) Harmonic(2)

Overtone is a musical term and it is not the same as harmonic.

An harmonic frequency is N times the fundamental, where as an
overtone merely holds a rational relationship to the fundamental
(3/2, 4/3 etc.)

The reason you can tell the difference between a violin and a
clarinet is that they have different overtone spectra.

A somewhat arbitrary subset of an intruments overtones are called
"formants" because they are the principal "formers" of the instruments
sound.

synchronous

  1. happening, existing or arising at precisely the same time
    2a) going on or operating together at exactly the same rate
    b) recurring together
  2. involving or indicating synchronism
  1. in physics, having the same period or having the same period and phase

Close, but no cigar.

If they merely have the same period/frequency, they are "syntonous",
but if the phase is also coincident, they are synchronous.

asynchronous
not synchronous, proceeding at its own pace;

But just because two signals are asynchronous, doesn't mean they cannot
have the same frequency or phase for a shorter or longer period of time,
it just means that there is no mechanism that makes it so.

The "asynchronous" serial line is actually anisochronous BTW.

No, each transmitter and receiver are anisochronous, but the
communication between them, and by language extension the actual
cable, is asynchronous.

This is interesting because overtone, when used to describe oscillation
modes in crystals, has a slightly different meaning.

When a crystal is made to oscillate on one of it's overtone
frequencies, the oscillation frequency is not exactly an harmonic of the
fundamental (or 1st harmonic) oscillation frequency.

This term is obviously overloaded and maybe we should not be using it.

Didier KO4BB

Poul-Henning Kamp wrote: > In message <20061226.125102.1723212534.cfmd@bredband.net>, Magnus Danielson wri > tes: > > >> harmonic(2) >> 1a) a tone in a harmonic series >> 2) in physics, a component frequency of a harmonic motion that is an integral >> multiple of the fundamental frequency >> >> overtone >> 1b) Harmonic(2) >> > > Overtone is a musical term and it is not the same as harmonic. > > An harmonic frequency is N times the fundamental, where as an > overtone merely holds a rational relationship to the fundamental > (3/2, 4/3 etc.) > > The reason you can tell the difference between a violin and a > clarinet is that they have different overtone spectra. > > A somewhat arbitrary subset of an intruments overtones are called > "formants" because they are the principal "formers" of the instruments > sound. > > >> synchronous >> 1) happening, existing or arising at precisely the same time >> 2a) going on or operating together at exactly the same rate >> b) recurring together >> 3) involving or indicating synchronism >> > > > >> 4) in physics, having the same period or having the same period and phase >> > > Close, but no cigar. > > If they merely have the same period/frequency, they are "syntonous", > but if the phase is also coincident, they are synchronous. > > >> asynchronous >> not synchronous, proceeding at its own pace; >> > > But just because two signals are asynchronous, doesn't mean they cannot > have the same frequency or phase for a shorter or longer period of time, > it just means that there is no mechanism that makes it so. > > >> The "asynchronous" serial line is actually anisochronous BTW. >> > > No, each transmitter and receiver are anisochronous, but the > communication between them, and by language extension the actual > cable, is asynchronous. > > This is interesting because overtone, when used to describe oscillation modes in crystals, has a slightly different meaning. When a crystal is made to oscillate on one of it's *overtone* frequencies, the oscillation frequency is not exactly an harmonic of the *fundamental* (or 1st harmonic) oscillation frequency. This term is obviously overloaded and maybe we should not be using it. Didier KO4BB
DB
Dr Bruce Griffiths
Tue, Dec 26, 2006 9:39 PM

Didier Juges wrote:

Dr Bruce Griffiths wrote:

Didier Juges wrote:

This discussion is fascinating, and as always it has prompted a number
of other questions for me.

I understand the sawtooth correction is provided to allow correction of
1 PPS timing errors when the processor clock is non-coherent with the
GPS signal (at least not intentionally), because the processor clock is
running at a finite frequency and provides granularity to the PPS
adjustment capability.

I wanted to know if this was available in my Trimble Thunderbolt GPSDO.

I have looked over the entire Thunderbolt manual and I have not seen a
mention about sawtooth correction, so I am not sure if it is available.

Then I realized the Thunderbolt is different from most GPSDO that have
been discussed, in that it uses the 10 MHz from the OCXO as the clock
for the GPS receiver, so in theory, the sawtooth correction should not
be necessary on this type of receiver since the processor clock and the
GPS signals are coherent (maybe I should not use that term...) when the
receiver is tracking.

Yet, the Thunderbolt spec for timing accuracy  on the 1 PPS output is
+/- 20 nS at 1 sigma, just like an ordinary GPSDO with stand alone
receiver and what would be about a 25 MHz clock (please confirm).

The Trimble software also displays values referred to as "Timing Output"
for the 1 PPS and the 10 MHz. The PPS value is currently hovering around
-1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x ppb. I am not
sure what that is based on. The manual simply refers to those as
"Estimate of UTC/GPS offset", but does not give any detail of how they
are computed or what they mean.

The Thunderbolt User's Manual says that the 1 PPS output is the OCXO's
10 MHz divided down. Yet, the block diagram shows the 1 PPS coming from
the CPU and support circuit, not directly coming from the 10 MHz. I
believe the 1 PPS is the 10 MHz divided down, except when the receiver
is doing jam-sync (after power up or recovery, when the GPS_PPS and
HW_PPS are far away from each other)

Going back to what Poul-Henning just said, I believe that for the
Thunderbolt, there actually may be three PPS signals:

GPS_PPS: what the receiver thinks the PPS should be (a theoretical signal)
HW_PPS: the best approximation of the GPS_PPS that the receiver can do
OCXO_PPS: the 10 MHz divided down

In jam-sync mode, the Thunderbolt forces the HW_PPS to within 100nS of
GPS_PPS (the closest it can get using software delays and programmable
dividers I guess, using a 10 MHz clock) without touching the OCXO, so in
that mode, HW_PPS and OCXO_PPS are obviously not the same. Once the two
are within 100 nS, the Thunderbolt switches to discipline mode and fine
tunes the OCXO to get the HW_PPS as close to GPS_PPS as possible (20nS 1
sigma by specification.) It would appear that the receiver should be
able to adjust its OCXO as close to GPS_PPS as the DAC will allow,
without the artificial limit of the processor clock period. So, there
should not be a need for sawtooth correction and there should be no
hanging bridges either.

Then, what are the "Timing Outputs"?

Does this make any sense?

Didier KO4BB


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Didier

Look at:
http://trl.trimble.com/docushare/dsweb/Get/Document-8425/

There is little evidence of sawtooth corrections in graph in the above
document, however the data were taken when SA was switched on.
To confuse matters further the document actually talks about pulse
position quantisation.

http://trl.trimble.com/docushare/dsweb/Get/Document-8424/
Shows differences between 2 Thunderbolts, again there is little evidence
of quantisation but data may have been smoothed.

Have you any reasonably stable crystal oscillator that can be used to
log the time interval between the Thunderbolt PPS output and the divided
down oscillator output?
Plotting these results should quickly show evidence of any sawtooth error.

As you say there shouldn't be any sawtooth error if the receiver timing
is derived from the 10MHz clock which is in turned locked to the PPS output.
However there are other sources of noise in a GPS timing receiver which
could, in an older receiver design like this, easily be as much as 20ns rms.

Bruce


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Bruce,

The Trimble app note you pointed to says the 1 PPS output is not
quantized, so I believe that means there should not be sawtooth error.

Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html

The last picture is the Thunderbolt PPS output against a free running HP
10811 divided by 256 with two 74F161.

The plot was obtained after removing steps, outliers (few, mostly due to
my acquisition software that puts some garbage in the output) and drift,
using Ulrich's Plotter program.

Since I was using the HP 5334, data is taken every 2 seconds.

How can I determine from this if sawtooth error is present?

Do I need to collect data differently?

Thanks

Didier

Note: my home brew GPIB data logging software records ambient
temperature at the same rate as the TI, which I believe would be a very
useful information. Unfortunately, I have not found yet how to display
both curves in a meaningful way using Plotter, so I have to strip the
temperature from the dataset before feeding it to Plotter. Also,
collecting data from two instruments causes my program to sometimes
insert garbage in the output file, which causes outliers and which I
need to fix before I put this software on my web page.


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Didier

The last paragraph should have started with As you say, in the absence
of noise, ......

In effect the 10MHz oscillator will track the smoothed PPS frequency.
However pulse by pulse the computed PPS signal timing will vary with
respect to the position of the smoothed PPS signal, consequently when
quantised there will be some variation of the timing of the PPS edge
with respect to the edge of the smoothed PPS pulse position. Thus some
evidence of quantisation will be seen in the PPS pulse output. The rate
at which such steps occur depends on the noise in the computed PPS
transition time. If the noise is significantly less than 100ns such
steps will be infrequent.
Even if the rms noise were around 100ns the resultant PPS timing error
plot will not exhibit the short term correlations evident in the Oncore
timing receiver sawtooth correction plot. The corresponding correction
plot should be more random.

The other complication is that the Thunderbolt noise spec was
derived/measured when SA was turned on. It may be significantly smaller
now that SA has been turned off.

Your plots show little or no evidence of any 50ns quantisation steps or
even 20ns steps.
This means that the internal receiver timing noise is significantly
smaller than the quantisation step.
The 20ns rms probably refers to the the receiver's internal
(prequantisation) timing noise in the presence of SA.

Bruce

Didier Juges wrote: > Dr Bruce Griffiths wrote: > >> Didier Juges wrote: >> >> >>> This discussion is fascinating, and as always it has prompted a number >>> of other questions for me. >>> >>> I understand the sawtooth correction is provided to allow correction of >>> 1 PPS timing errors when the processor clock is non-coherent with the >>> GPS signal (at least not intentionally), because the processor clock is >>> running at a finite frequency and provides granularity to the PPS >>> adjustment capability. >>> >>> I wanted to know if this was available in my Trimble Thunderbolt GPSDO. >>> >>> I have looked over the entire Thunderbolt manual and I have not seen a >>> mention about sawtooth correction, so I am not sure if it is available. >>> >>> Then I realized the Thunderbolt is different from most GPSDO that have >>> been discussed, in that it uses the 10 MHz from the OCXO as the clock >>> for the GPS receiver, so in theory, the sawtooth correction should not >>> be necessary on this type of receiver since the processor clock and the >>> GPS signals are coherent (maybe I should not use that term...) when the >>> receiver is tracking. >>> >>> Yet, the Thunderbolt spec for timing accuracy on the 1 PPS output is >>> +/- 20 nS at 1 sigma, just like an *ordinary* GPSDO with stand alone >>> receiver and what would be about a 25 MHz clock (please confirm). >>> >>> The Trimble software also displays values referred to as "Timing Output" >>> for the 1 PPS and the 10 MHz. The PPS value is currently hovering around >>> -1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x ppb. I am not >>> sure what that is based on. The manual simply refers to those as >>> "Estimate of UTC/GPS offset", but does not give any detail of how they >>> are computed or what they mean. >>> >>> The Thunderbolt User's Manual says that the 1 PPS output is the OCXO's >>> 10 MHz divided down. Yet, the block diagram shows the 1 PPS coming from >>> the CPU and support circuit, not directly coming from the 10 MHz. I >>> believe the 1 PPS is the 10 MHz divided down, except when the receiver >>> is doing jam-sync (after power up or recovery, when the GPS_PPS and >>> HW_PPS are far away from each other) >>> >>> Going back to what Poul-Henning just said, I believe that for the >>> Thunderbolt, there actually may be three PPS signals: >>> >>> GPS_PPS: what the receiver thinks the PPS should be (a theoretical signal) >>> HW_PPS: the best approximation of the GPS_PPS that the receiver can do >>> OCXO_PPS: the 10 MHz divided down >>> >>> In jam-sync mode, the Thunderbolt forces the HW_PPS to within 100nS of >>> GPS_PPS (the closest it can get using software delays and programmable >>> dividers I guess, using a 10 MHz clock) without touching the OCXO, so in >>> that mode, HW_PPS and OCXO_PPS are obviously not the same. Once the two >>> are within 100 nS, the Thunderbolt switches to discipline mode and fine >>> tunes the OCXO to get the HW_PPS as close to GPS_PPS as possible (20nS 1 >>> sigma by specification.) It would appear that the receiver should be >>> able to adjust its OCXO as close to GPS_PPS as the DAC will allow, >>> without the artificial limit of the processor clock period. So, there >>> should not be a need for sawtooth correction and there should be no >>> hanging bridges either. >>> >>> Then, what are the "Timing Outputs"? >>> >>> Does this make any sense? >>> >>> Didier KO4BB >>> >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list >>> time-nuts@febo.com >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >>> >>> >>> >> Didier >> >> Look at: >> http://trl.trimble.com/docushare/dsweb/Get/Document-8425/ >> >> There is little evidence of sawtooth corrections in graph in the above >> document, however the data were taken when SA was switched on. >> To confuse matters further the document actually talks about pulse >> position quantisation. >> >> http://trl.trimble.com/docushare/dsweb/Get/Document-8424/ >> Shows differences between 2 Thunderbolts, again there is little evidence >> of quantisation but data may have been smoothed. >> >> Have you any reasonably stable crystal oscillator that can be used to >> log the time interval between the Thunderbolt PPS output and the divided >> down oscillator output? >> Plotting these results should quickly show evidence of any sawtooth error. >> >> As you say there shouldn't be any sawtooth error if the receiver timing >> is derived from the 10MHz clock which is in turned locked to the PPS output. >> However there are other sources of noise in a GPS timing receiver which >> could, in an older receiver design like this, easily be as much as 20ns rms. >> >> Bruce >> >> _______________________________________________ >> time-nuts mailing list >> time-nuts@febo.com >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> > Bruce, > > The Trimble app note you pointed to says the 1 PPS output is not > quantized, so I believe that means there should not be sawtooth error. > > Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html > > The last picture is the Thunderbolt PPS output against a free running HP > 10811 divided by 256 with two 74F161. > > The plot was obtained after removing steps, outliers (few, mostly due to > my acquisition software that puts some garbage in the output) and drift, > using Ulrich's Plotter program. > > Since I was using the HP 5334, data is taken every 2 seconds. > > How can I determine from this if sawtooth error is present? > > Do I need to collect data differently? > > Thanks > > Didier > > Note: my home brew GPIB data logging software records ambient > temperature at the same rate as the TI, which I believe would be a very > useful information. Unfortunately, I have not found yet how to display > both curves in a meaningful way using Plotter, so I have to strip the > temperature from the dataset before feeding it to Plotter. Also, > collecting data from two instruments causes my program to sometimes > insert garbage in the output file, which causes outliers and which I > need to fix before I put this software on my web page. > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > Didier The last paragraph should have started with As you say, in the absence of noise, ...... In effect the 10MHz oscillator will track the smoothed PPS frequency. However pulse by pulse the computed PPS signal timing will vary with respect to the position of the smoothed PPS signal, consequently when quantised there will be some variation of the timing of the PPS edge with respect to the edge of the smoothed PPS pulse position. Thus some evidence of quantisation will be seen in the PPS pulse output. The rate at which such steps occur depends on the noise in the computed PPS transition time. If the noise is significantly less than 100ns such steps will be infrequent. Even if the rms noise were around 100ns the resultant PPS timing error plot will not exhibit the short term correlations evident in the Oncore timing receiver sawtooth correction plot. The corresponding correction plot should be more random. The other complication is that the Thunderbolt noise spec was derived/measured when SA was turned on. It may be significantly smaller now that SA has been turned off. Your plots show little or no evidence of any 50ns quantisation steps or even 20ns steps. This means that the internal receiver timing noise is significantly smaller than the quantisation step. The 20ns rms probably refers to the the receiver's internal (prequantisation) timing noise in the presence of SA. Bruce
DJ
Didier Juges
Tue, Dec 26, 2006 10:14 PM

Dr Bruce Griffiths wrote:

Didier Juges wrote:

Dr Bruce Griffiths wrote:

Didier Juges wrote:

This discussion is fascinating, and as always it has prompted a number
of other questions for me.

I understand the sawtooth correction is provided to allow correction of
1 PPS timing errors when the processor clock is non-coherent with the
GPS signal (at least not intentionally), because the processor clock is
running at a finite frequency and provides granularity to the PPS
adjustment capability.

I wanted to know if this was available in my Trimble Thunderbolt GPSDO.

I have looked over the entire Thunderbolt manual and I have not seen a
mention about sawtooth correction, so I am not sure if it is available.

Then I realized the Thunderbolt is different from most GPSDO that have
been discussed, in that it uses the 10 MHz from the OCXO as the clock
for the GPS receiver, so in theory, the sawtooth correction should not
be necessary on this type of receiver since the processor clock and the
GPS signals are coherent (maybe I should not use that term...) when the
receiver is tracking.

Yet, the Thunderbolt spec for timing accuracy  on the 1 PPS output is
+/- 20 nS at 1 sigma, just like an ordinary GPSDO with stand alone
receiver and what would be about a 25 MHz clock (please confirm).

The Trimble software also displays values referred to as "Timing Output"
for the 1 PPS and the 10 MHz. The PPS value is currently hovering around
-1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x ppb. I am not
sure what that is based on. The manual simply refers to those as
"Estimate of UTC/GPS offset", but does not give any detail of how they
are computed or what they mean.

The Thunderbolt User's Manual says that the 1 PPS output is the OCXO's
10 MHz divided down. Yet, the block diagram shows the 1 PPS coming from
the CPU and support circuit, not directly coming from the 10 MHz. I
believe the 1 PPS is the 10 MHz divided down, except when the receiver
is doing jam-sync (after power up or recovery, when the GPS_PPS and
HW_PPS are far away from each other)

Going back to what Poul-Henning just said, I believe that for the
Thunderbolt, there actually may be three PPS signals:

GPS_PPS: what the receiver thinks the PPS should be (a theoretical signal)
HW_PPS: the best approximation of the GPS_PPS that the receiver can do
OCXO_PPS: the 10 MHz divided down

In jam-sync mode, the Thunderbolt forces the HW_PPS to within 100nS of
GPS_PPS (the closest it can get using software delays and programmable
dividers I guess, using a 10 MHz clock) without touching the OCXO, so in
that mode, HW_PPS and OCXO_PPS are obviously not the same. Once the two
are within 100 nS, the Thunderbolt switches to discipline mode and fine
tunes the OCXO to get the HW_PPS as close to GPS_PPS as possible (20nS 1
sigma by specification.) It would appear that the receiver should be
able to adjust its OCXO as close to GPS_PPS as the DAC will allow,
without the artificial limit of the processor clock period. So, there
should not be a need for sawtooth correction and there should be no
hanging bridges either.

Then, what are the "Timing Outputs"?

Does this make any sense?

Didier KO4BB


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Didier

Look at:
http://trl.trimble.com/docushare/dsweb/Get/Document-8425/

There is little evidence of sawtooth corrections in graph in the above
document, however the data were taken when SA was switched on.
To confuse matters further the document actually talks about pulse
position quantisation.

http://trl.trimble.com/docushare/dsweb/Get/Document-8424/
Shows differences between 2 Thunderbolts, again there is little evidence
of quantisation but data may have been smoothed.

Have you any reasonably stable crystal oscillator that can be used to
log the time interval between the Thunderbolt PPS output and the divided
down oscillator output?
Plotting these results should quickly show evidence of any sawtooth error.

As you say there shouldn't be any sawtooth error if the receiver timing
is derived from the 10MHz clock which is in turned locked to the PPS output.
However there are other sources of noise in a GPS timing receiver which
could, in an older receiver design like this, easily be as much as 20ns rms.

Bruce


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Bruce,

The Trimble app note you pointed to says the 1 PPS output is not
quantized, so I believe that means there should not be sawtooth error.

Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html

The last picture is the Thunderbolt PPS output against a free running HP
10811 divided by 256 with two 74F161.

The plot was obtained after removing steps, outliers (few, mostly due to
my acquisition software that puts some garbage in the output) and drift,
using Ulrich's Plotter program.

Since I was using the HP 5334, data is taken every 2 seconds.

How can I determine from this if sawtooth error is present?

Do I need to collect data differently?

Thanks

Didier

Note: my home brew GPIB data logging software records ambient
temperature at the same rate as the TI, which I believe would be a very
useful information. Unfortunately, I have not found yet how to display
both curves in a meaningful way using Plotter, so I have to strip the
temperature from the dataset before feeding it to Plotter. Also,
collecting data from two instruments causes my program to sometimes
insert garbage in the output file, which causes outliers and which I
need to fix before I put this software on my web page.


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Didier

The last paragraph should have started with As you say, in the absence
of noise, ......

In effect the 10MHz oscillator will track the smoothed PPS frequency.
However pulse by pulse the computed PPS signal timing will vary with
respect to the position of the smoothed PPS signal, consequently when
quantised there will be some variation of the timing of the PPS edge
with respect to the edge of the smoothed PPS pulse position. Thus some
evidence of quantisation will be seen in the PPS pulse output. The rate
at which such steps occur depends on the noise in the computed PPS
transition time. If the noise is significantly less than 100ns such
steps will be infrequent.
Even if the rms noise were around 100ns the resultant PPS timing error
plot will not exhibit the short term correlations evident in the Oncore
timing receiver sawtooth correction plot. The corresponding correction
plot should be more random.

The other complication is that the Thunderbolt noise spec was
derived/measured when SA was turned on. It may be significantly smaller
now that SA has been turned off.

Your plots show little or no evidence of any 50ns quantisation steps or
even 20ns steps.
This means that the internal receiver timing noise is significantly
smaller than the quantisation step.
The 20ns rms probably refers to the the receiver's internal
(prequantisation) timing noise in the presence of SA.

Bruce

Thank you Bruce, that's very encouraging.

Regarding the jam-sync mode, I am not sure how much the GPS_PPS and
HW_PPS need to be separated by before the jam-sync mode in engaged. I
believe it takes more than one pulse because typically this mode is only
used when recovering from holdover or after startup, so it should not be
engaged as long as the receiver has a valid fix, even when satellites
come in and out. So I would not expect to see phase jumps unless the GPS
signal went completely away. I need to write another logging program to
log the serial data from the Thunderbolt and see how often, if at all,
it goes into holdover with my currently marginal antenna situation. I
have not seen it happen since I moved the antenna in its current
location though.

When reading the data sheet for the Thunderbolt, and reading all the
pitfalls associated with non-integrated GPSDO designs using stand alone
GPS receivers, such as sawtooth correction and quantization noise, it
seems like the integrated approach used in the Thunderbolt is the best
way to go. Not only it is cheaper and simpler, and therefore should be
more reliable, but it avoids an entire class of problems and should
perform better, everything else being equal (receiver sensitivity and
internal noise, OCXO quality). In this case, simplicity goes with better
performance, which is not always the case.

Yet, I am surprised that so many of the OEM timing receiver solutions
use the conventional approach. For instance, the Lucent receiver John
just bought seems to have a discrete, independent GPS receiver (the
darker board on top), and many companies still build stand alone GPS
receivers specifically for timing application without embedding the
GPSDO logic and an OCXO. That does not seem to make much sense to me.

In the mean time, I keep forgetting the Thunderbolt specs were written
when SA was turned on, and therefore better performance should be
expected today from the same hardware.

Tom said he was planning a comparison test session of various GPSDO
designs in January or February I believe. If he does not have one, I'll
be glad to send him my Thunderbolt for comparison. Tom, if you are
interested, please speak up.

Thanks again.

Didier

Dr Bruce Griffiths wrote: > Didier Juges wrote: > >> Dr Bruce Griffiths wrote: >> >> >>> Didier Juges wrote: >>> >>> >>> >>>> This discussion is fascinating, and as always it has prompted a number >>>> of other questions for me. >>>> >>>> I understand the sawtooth correction is provided to allow correction of >>>> 1 PPS timing errors when the processor clock is non-coherent with the >>>> GPS signal (at least not intentionally), because the processor clock is >>>> running at a finite frequency and provides granularity to the PPS >>>> adjustment capability. >>>> >>>> I wanted to know if this was available in my Trimble Thunderbolt GPSDO. >>>> >>>> I have looked over the entire Thunderbolt manual and I have not seen a >>>> mention about sawtooth correction, so I am not sure if it is available. >>>> >>>> Then I realized the Thunderbolt is different from most GPSDO that have >>>> been discussed, in that it uses the 10 MHz from the OCXO as the clock >>>> for the GPS receiver, so in theory, the sawtooth correction should not >>>> be necessary on this type of receiver since the processor clock and the >>>> GPS signals are coherent (maybe I should not use that term...) when the >>>> receiver is tracking. >>>> >>>> Yet, the Thunderbolt spec for timing accuracy on the 1 PPS output is >>>> +/- 20 nS at 1 sigma, just like an *ordinary* GPSDO with stand alone >>>> receiver and what would be about a 25 MHz clock (please confirm). >>>> >>>> The Trimble software also displays values referred to as "Timing Output" >>>> for the 1 PPS and the 10 MHz. The PPS value is currently hovering around >>>> -1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x ppb. I am not >>>> sure what that is based on. The manual simply refers to those as >>>> "Estimate of UTC/GPS offset", but does not give any detail of how they >>>> are computed or what they mean. >>>> >>>> The Thunderbolt User's Manual says that the 1 PPS output is the OCXO's >>>> 10 MHz divided down. Yet, the block diagram shows the 1 PPS coming from >>>> the CPU and support circuit, not directly coming from the 10 MHz. I >>>> believe the 1 PPS is the 10 MHz divided down, except when the receiver >>>> is doing jam-sync (after power up or recovery, when the GPS_PPS and >>>> HW_PPS are far away from each other) >>>> >>>> Going back to what Poul-Henning just said, I believe that for the >>>> Thunderbolt, there actually may be three PPS signals: >>>> >>>> GPS_PPS: what the receiver thinks the PPS should be (a theoretical signal) >>>> HW_PPS: the best approximation of the GPS_PPS that the receiver can do >>>> OCXO_PPS: the 10 MHz divided down >>>> >>>> In jam-sync mode, the Thunderbolt forces the HW_PPS to within 100nS of >>>> GPS_PPS (the closest it can get using software delays and programmable >>>> dividers I guess, using a 10 MHz clock) without touching the OCXO, so in >>>> that mode, HW_PPS and OCXO_PPS are obviously not the same. Once the two >>>> are within 100 nS, the Thunderbolt switches to discipline mode and fine >>>> tunes the OCXO to get the HW_PPS as close to GPS_PPS as possible (20nS 1 >>>> sigma by specification.) It would appear that the receiver should be >>>> able to adjust its OCXO as close to GPS_PPS as the DAC will allow, >>>> without the artificial limit of the processor clock period. So, there >>>> should not be a need for sawtooth correction and there should be no >>>> hanging bridges either. >>>> >>>> Then, what are the "Timing Outputs"? >>>> >>>> Does this make any sense? >>>> >>>> Didier KO4BB >>>> >>>> >>>> >>>> >>>> _______________________________________________ >>>> time-nuts mailing list >>>> time-nuts@febo.com >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> >>>> >>>> >>>> >>>> >>> Didier >>> >>> Look at: >>> http://trl.trimble.com/docushare/dsweb/Get/Document-8425/ >>> >>> There is little evidence of sawtooth corrections in graph in the above >>> document, however the data were taken when SA was switched on. >>> To confuse matters further the document actually talks about pulse >>> position quantisation. >>> >>> http://trl.trimble.com/docushare/dsweb/Get/Document-8424/ >>> Shows differences between 2 Thunderbolts, again there is little evidence >>> of quantisation but data may have been smoothed. >>> >>> Have you any reasonably stable crystal oscillator that can be used to >>> log the time interval between the Thunderbolt PPS output and the divided >>> down oscillator output? >>> Plotting these results should quickly show evidence of any sawtooth error. >>> >>> As you say there shouldn't be any sawtooth error if the receiver timing >>> is derived from the 10MHz clock which is in turned locked to the PPS output. >>> However there are other sources of noise in a GPS timing receiver which >>> could, in an older receiver design like this, easily be as much as 20ns rms. >>> >>> Bruce >>> >>> _______________________________________________ >>> time-nuts mailing list >>> time-nuts@febo.com >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> >>> >>> >>> >> Bruce, >> >> The Trimble app note you pointed to says the 1 PPS output is not >> quantized, so I believe that means there should not be sawtooth error. >> >> Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html >> >> The last picture is the Thunderbolt PPS output against a free running HP >> 10811 divided by 256 with two 74F161. >> >> The plot was obtained after removing steps, outliers (few, mostly due to >> my acquisition software that puts some garbage in the output) and drift, >> using Ulrich's Plotter program. >> >> Since I was using the HP 5334, data is taken every 2 seconds. >> >> How can I determine from this if sawtooth error is present? >> >> Do I need to collect data differently? >> >> Thanks >> >> Didier >> >> Note: my home brew GPIB data logging software records ambient >> temperature at the same rate as the TI, which I believe would be a very >> useful information. Unfortunately, I have not found yet how to display >> both curves in a meaningful way using Plotter, so I have to strip the >> temperature from the dataset before feeding it to Plotter. Also, >> collecting data from two instruments causes my program to sometimes >> insert garbage in the output file, which causes outliers and which I >> need to fix before I put this software on my web page. >> >> _______________________________________________ >> time-nuts mailing list >> time-nuts@febo.com >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> >> >> > Didier > > The last paragraph should have started with As you say, in the absence > of noise, ...... > > In effect the 10MHz oscillator will track the smoothed PPS frequency. > However pulse by pulse the computed PPS signal timing will vary with > respect to the position of the smoothed PPS signal, consequently when > quantised there will be some variation of the timing of the PPS edge > with respect to the edge of the smoothed PPS pulse position. Thus some > evidence of quantisation will be seen in the PPS pulse output. The rate > at which such steps occur depends on the noise in the computed PPS > transition time. If the noise is significantly less than 100ns such > steps will be infrequent. > Even if the rms noise were around 100ns the resultant PPS timing error > plot will not exhibit the short term correlations evident in the Oncore > timing receiver sawtooth correction plot. The corresponding correction > plot should be more random. > > The other complication is that the Thunderbolt noise spec was > derived/measured when SA was turned on. It may be significantly smaller > now that SA has been turned off. > > Your plots show little or no evidence of any 50ns quantisation steps or > even 20ns steps. > This means that the internal receiver timing noise is significantly > smaller than the quantisation step. > The 20ns rms probably refers to the the receiver's internal > (prequantisation) timing noise in the presence of SA. > > Bruce > Thank you Bruce, that's very encouraging. Regarding the jam-sync mode, I am not sure how much the GPS_PPS and HW_PPS need to be separated by before the jam-sync mode in engaged. I believe it takes more than one pulse because typically this mode is only used when recovering from holdover or after startup, so it should not be engaged as long as the receiver has a valid fix, even when satellites come in and out. So I would not expect to see phase jumps unless the GPS signal went completely away. I need to write another logging program to log the serial data from the Thunderbolt and see how often, if at all, it goes into holdover with my currently marginal antenna situation. I have not seen it happen since I moved the antenna in its current location though. When reading the data sheet for the Thunderbolt, and reading all the pitfalls associated with non-integrated GPSDO designs using stand alone GPS receivers, such as sawtooth correction and quantization noise, it seems like the integrated approach used in the Thunderbolt is the best way to go. Not only it is cheaper and simpler, and therefore should be more reliable, but it avoids an entire class of problems and should perform better, everything else being equal (receiver sensitivity and internal noise, OCXO quality). In this case, simplicity goes with better performance, which is not always the case. Yet, I am surprised that so many of the OEM timing receiver solutions use the *conventional* approach. For instance, the Lucent receiver John just bought seems to have a discrete, independent GPS receiver (the darker board on top), and many companies still build stand alone GPS receivers specifically for timing application without embedding the GPSDO logic and an OCXO. That does not seem to make much sense to me. In the mean time, I keep forgetting the Thunderbolt specs were written when SA was turned on, and therefore better performance should be expected today from the same hardware. Tom said he was planning a comparison test session of various GPSDO designs in January or February I believe. If he does not have one, I'll be glad to send him my Thunderbolt for comparison. Tom, if you are interested, please speak up. Thanks again. Didier
MD
Magnus Danielson
Tue, Dec 26, 2006 10:23 PM

From: Dr Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] TIC resolution impact on GPSDO's performance
Date: Wed, 27 Dec 2006 01:15:48 +1300
Message-ID: 45911274.5030605@xtra.co.nz

Magnus Danielson wrote:

From: "Poul-Henning Kamp" phk@phk.freebsd.dk
Subject: Re: [time-nuts] TIC resolution impact on GPSDO's performance
Date: Tue, 26 Dec 2006 08:14:12 +0000
Message-ID: 36958.1167120852@critter.freebsd.dk

In message 4590BB56.5070809@xtra.co.nz, Dr Bruce Griffiths writes:

Poul-Henning Kamp wrote:

I think what you are trying to express is that the frequency from
the internal Xtal (at times) is in an overtone of the 1Hz PPS, which
gives rise to hanging bridges.

Poul-Henning

Surely you mean harmonic? An overtone is not necessarily a harmonic.

I meant overtone, becuase they are not in any harmonic relationship,
the XO is wandering around whereas the (ideal) PPS is, supposedly,
rock stable.

The only real way to describe it is to say that the XO and the PPS is
asynchronous to each other. Those who has freshen up on their greek for
technicians will know that this means "not the same clock" which is quite
accuratly what we have. ITU-T Rec. G.700 has a few interesting sections in it.

The use of words such as harmonic or overtone should not be used since they
is normally used to describe the various frequencies within one signal where
as words as synchronous and asynchronous is to be used mainly for pair of
signals and describes their relative timing. Naturally, for a pair of signals
to be synchronous does not require them to have the same frequency, just a
fixed ratio between those frequencies.

It does happends that a single signal is said to be synchronous, but the
wording which should have been used is isochronous (same clock). Then we have
the lovely words of plesiochronous and mesochronous. :)

Cheers,
Magnus


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Magnus

Bruce,

So the correct description of the cause of hanging bridges is something
like?

As the frequency of the clock which controls the timing of the receiver
PPS pulse output drifts and approaches pleisiosynchronism with the
"true" frequency of the PPS signal, the rate of change of the PPS timing
error due to quantisation decreases reaching zero should the clock
achieve synchronism with the "true" PPS frequency.
Thus hanging bridges occur when the clock frequency approaches
pleisiosynchronism with the "true" PPS signal frequency and then drifts
away from pleisiosynchronism with the PPS signal..

This is much better even if it is not more readable, but that is in the
beholders eye.

However, there is one thing which may not be all that clear. For each PPS pulse
to be generated, the firmware needs to decide which clock-transition will best
approximate the internal time solution. This means that it continously makes a
frequency justification. PDH guys would call this bit justification and SDH
guys would call this pointer justification. Justifications of this kind is
known for one thing, being bad of relasing detailed phase information. The lack
of transition causes the worstcase quantization error to manifest itself.
The high frequency error causes many justifications and when they are smoothed
out they give a higher degree of phase information. Having to rely on such a
justification mechanism isn't very nice. The negative sawtooth error output
really removes much of the problem and especially relating to the hanging
bridge problem becomes less of an issue by using the information. However, the
negative sawtooth error signal is limited in resolution so on modern receivers
the full potential of the receiver frontend.

When the frequency of the clock controlling the timing of the receiver
PPS pulse output is synchronous with the PPS frequency, then neither
sawtooth corrections or hanging bridges occur, however there may be a
fixed offset of upto 1/2 a clock cycle between the "correct" and the
actual leading edge of the receiver PPS output pulse.

The hanging bridge of the sawtooth correction is much smaller than that of the
PPS only. A hanging bridge of the PPS regulation can still output phase errors
and thus reduce the effective quantization error. The effect is there thought,
and for a shorter time.

Please correct and simplify these statements as required.

A certain NZ product has my main attention right now. They have some problem
with a ringmod or some ringing... they use old tools and old magic and takes a
hell of a time to acheive. Back to the TV.

But yes, it could probably be better explained.

The only problem that I can see is we still need to define how close to
synchronism the timing clock frequency and the "true" PPS signal
frequency system must be to be considered pleisiosynchronous.

As always, what degree of synchronousity is meaningfull for us?

Also, do we really have to care? Is this really how this problem should be
solved? Just because this is how receivers traditionally have indicated time,
is this really the best solution to compare a clock, via a for our purposes a
really poor clock? I think not.

Cheers,
Magnus

From: Dr Bruce Griffiths <bruce.griffiths@xtra.co.nz> Subject: Re: [time-nuts] TIC resolution impact on GPSDO's performance Date: Wed, 27 Dec 2006 01:15:48 +1300 Message-ID: <45911274.5030605@xtra.co.nz> > Magnus Danielson wrote: > > From: "Poul-Henning Kamp" <phk@phk.freebsd.dk> > > Subject: Re: [time-nuts] TIC resolution impact on GPSDO's performance > > Date: Tue, 26 Dec 2006 08:14:12 +0000 > > Message-ID: <36958.1167120852@critter.freebsd.dk> > > > > > >> In message <4590BB56.5070809@xtra.co.nz>, Dr Bruce Griffiths writes: > >> > >>> Poul-Henning Kamp wrote: > >>> > >>> > >>>> I think what you are trying to express is that the frequency from > >>>> the internal Xtal (at times) is in an overtone of the 1Hz PPS, which > >>>> gives rise to hanging bridges. > >>>> > >>>> > >>>> > >>> Poul-Henning > >>> > >>> Surely you mean harmonic? An overtone is not necessarily a harmonic. > >>> > >> I meant overtone, becuase they are not in any harmonic relationship, > >> the XO is wandering around whereas the (ideal) PPS is, supposedly, > >> rock stable. > >> > > > > The only real way to describe it is to say that the XO and the PPS is > > asynchronous to each other. Those who has freshen up on their greek for > > technicians will know that this means "not the same clock" which is quite > > accuratly what we have. ITU-T Rec. G.700 has a few interesting sections in it. > > > > The use of words such as harmonic or overtone should not be used since they > > is normally used to describe the various frequencies within one signal where > > as words as synchronous and asynchronous is to be used mainly for pair of > > signals and describes their relative timing. Naturally, for a pair of signals > > to be synchronous does not require them to have the same frequency, just a > > fixed ratio between those frequencies. > > > > It does happends that a single signal is said to be synchronous, but the > > wording which should have been used is isochronous (same clock). Then we have > > the lovely words of plesiochronous and mesochronous. :) > > > > Cheers, > > Magnus > > > > _______________________________________________ > > time-nuts mailing list > > time-nuts@febo.com > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > > > Magnus Bruce, > So the correct description of the cause of hanging bridges is something > like? > > As the frequency of the clock which controls the timing of the receiver > PPS pulse output drifts and approaches pleisiosynchronism with the > "true" frequency of the PPS signal, the rate of change of the PPS timing > error due to quantisation decreases reaching zero should the clock > achieve synchronism with the "true" PPS frequency. > Thus hanging bridges occur when the clock frequency approaches > pleisiosynchronism with the "true" PPS signal frequency and then drifts > away from pleisiosynchronism with the PPS signal.. This is much better even if it is not more readable, but that is in the beholders eye. However, there is one thing which may not be all that clear. For each PPS pulse to be generated, the firmware needs to decide which clock-transition will best approximate the internal time solution. This means that it continously makes a frequency justification. PDH guys would call this bit justification and SDH guys would call this pointer justification. Justifications of this kind is known for one thing, being bad of relasing detailed phase information. The lack of transition causes the worstcase quantization error to manifest itself. The high frequency error causes many justifications and when they are smoothed out they give a higher degree of phase information. Having to rely on such a justification mechanism isn't very nice. The negative sawtooth error output really removes much of the problem and especially relating to the hanging bridge problem becomes less of an issue by using the information. However, the negative sawtooth error signal is limited in resolution so on modern receivers the full potential of the receiver frontend. > When the frequency of the clock controlling the timing of the receiver > PPS pulse output is synchronous with the PPS frequency, then neither > sawtooth corrections or hanging bridges occur, however there may be a > fixed offset of upto 1/2 a clock cycle between the "correct" and the > actual leading edge of the receiver PPS output pulse. The hanging bridge of the sawtooth correction is much smaller than that of the PPS only. A hanging bridge of the PPS regulation can still output phase errors and thus reduce the effective quantization error. The effect is there thought, and for a shorter time. > Please correct and simplify these statements as required. A certain NZ product has my main attention right now. They have some problem with a ringmod or some ringing... they use old tools and old magic and takes a hell of a time to acheive. Back to the TV. But yes, it could probably be better explained. > The only problem that I can see is we still need to define how close to > synchronism the timing clock frequency and the "true" PPS signal > frequency system must be to be considered pleisiosynchronous. As always, what degree of synchronousity is meaningfull for us? Also, do we really *have* to care? Is this *really* how this problem should be solved? Just because this is how receivers traditionally have indicated time, is this really the best solution to compare a clock, via a for our purposes a really poor clock? I think not. Cheers, Magnus
DB
Dr Bruce Griffiths
Wed, Dec 27, 2006 3:32 AM

Magnus

Amended to substitute plesiochronous and plesiochronism for the
incorrect words I used previously.

As the frequency of the clock which controls the timing of the receiver
PPS pulse output drifts and approaches plesiochronism with the
"true" frequency of the PPS signal, the rate of change of the PPS timing
error due to quantisation decreases reaching zero should the clock
achieve synchronism with the "true" PPS frequency.
Thus hanging bridges occur when the clock frequency approaches
plesiochronism with the "true" PPS signal frequency and then drifts
away from plesiochronism with the PPS signal..

When the frequency of the clock controlling the timing of the receiver
PPS pulse output is synchronous with the PPS frequency, then neither
sawtooth corrections or hanging bridges occur, however there may be a
fixed offset of up to 1/2 a clock cycle between the "correct" and the
actual leading edge of the receiver PPS output pulse.

Please correct and simplify these statements as required.

The only problem that I can see is we still need to define how close to
synchronism the timing clock frequency and the "true" PPS signal
frequency system must be to be considered plesiochronous.

Bruce

Magnus Amended to substitute plesiochronous and plesiochronism for the incorrect words I used previously. As the frequency of the clock which controls the timing of the receiver PPS pulse output drifts and approaches plesiochronism with the "true" frequency of the PPS signal, the rate of change of the PPS timing error due to quantisation decreases reaching zero should the clock achieve synchronism with the "true" PPS frequency. Thus hanging bridges occur when the clock frequency approaches plesiochronism with the "true" PPS signal frequency and then drifts away from plesiochronism with the PPS signal.. When the frequency of the clock controlling the timing of the receiver PPS pulse output is synchronous with the PPS frequency, then neither sawtooth corrections or hanging bridges occur, however there may be a fixed offset of up to 1/2 a clock cycle between the "correct" and the actual leading edge of the receiver PPS output pulse. Please correct and simplify these statements as required. The only problem that I can see is we still need to define how close to synchronism the timing clock frequency and the "true" PPS signal frequency system must be to be considered plesiochronous. Bruce
DI
David I. Emery
Wed, Dec 27, 2006 6:21 AM

On Tue, Dec 26, 2006 at 04:14:45PM -0600, Didier Juges wrote:

When reading the data sheet for the Thunderbolt, and reading all the
pitfalls associated with non-integrated GPSDO designs using stand alone
GPS receivers, such as sawtooth correction and quantization noise, it
seems like the integrated approach used in the Thunderbolt is the best
way to go. Not only it is cheaper and simpler, and therefore should be
more reliable, but it avoids an entire class of problems and should
perform better, everything else being equal (receiver sensitivity and
internal noise, OCXO quality). In this case, simplicity goes with better
performance, which is not always the case.

It is my understanding that the Motorola Oncore timing receiver

modules (UT,VP and onwards to the M12+ family) can be ordered in a
version that accepts an EXTERNAL 10 mhz rather than using an internal
crystal for timing.  At least I think it is a 10 mhz input, but I am
quite sure they can be ordered in a version that does not generate
timing or frequency from an internal xtal oscillator.

And if my presumption is correct that the Trimble Thunderbolt

hardware is either identical to or very similar to the HP/Symmetricom
58540A hardware (and both OEMed from a company in Japan) I think you
will find that they do use a Motorola GPS timing receiver in external
clocked mode using a clock derived from the 10 mhz OXCO.  I beleive my
58540As do anyway.

Yet, I am surprised that so many of the OEM timing receiver solutions
use the conventional approach. For instance, the Lucent receiver John
just bought seems to have a discrete, independent GPS receiver (the
darker board on top), and many companies still build stand alone GPS
receivers specifically for timing application without embedding the
GPSDO logic and an OCXO. That does not seem to make much sense to me.

I have ordered a Lucent RFTG-M-XO box but it has yet to arrive -

however, looking at the photos published today, it does not seem clear
they don't use a GPS receiver with an external clock.  Not easy to tell
since the guts of the Motorola module are under a shield can.

There is certainly no reason to integrate a GPS receiver with

the OXCO PLL and status electronics if one can buy an OEM one that takes
an external clock for less money (and hastle over firmware development
and licensing for the GPS side).  GPSDOs are a small market compared to
the overall market for OEM GPS solutions and there are economies of
scale involved.

And as a final note - a Datum 9390 box I have that dates to

beginning of time (GPS time that is) used a Rb derived clock for the
antique OEM Trimble GPS receiver board it uses so that approach has been
around from the early days...

--
Dave Emery N1PRE,  die@dieconsulting.com  DIE Consulting, Weston, Mass 02493
"An empty zombie mind with a forlorn barely readable weatherbeaten
'For Rent' sign still vainly flapping outside on the weed encrusted pole - in
celebration of what could have been, but wasn't and is not to be now either."

On Tue, Dec 26, 2006 at 04:14:45PM -0600, Didier Juges wrote: > When reading the data sheet for the Thunderbolt, and reading all the > pitfalls associated with non-integrated GPSDO designs using stand alone > GPS receivers, such as sawtooth correction and quantization noise, it > seems like the integrated approach used in the Thunderbolt is the best > way to go. Not only it is cheaper and simpler, and therefore should be > more reliable, but it avoids an entire class of problems and should > perform better, everything else being equal (receiver sensitivity and > internal noise, OCXO quality). In this case, simplicity goes with better > performance, which is not always the case. It is my understanding that the Motorola Oncore timing receiver modules (UT,VP and onwards to the M12+ family) can be ordered in a version that accepts an EXTERNAL 10 mhz rather than using an internal crystal for timing. At least I think it is a 10 mhz input, but I am quite sure they can be ordered in a version that does not generate timing or frequency from an internal xtal oscillator. And if my presumption is correct that the Trimble Thunderbolt hardware is either identical to or very similar to the HP/Symmetricom 58540A hardware (and both OEMed from a company in Japan) I think you will find that they do use a Motorola GPS timing receiver in external clocked mode using a clock derived from the 10 mhz OXCO. I beleive my 58540As do anyway. > Yet, I am surprised that so many of the OEM timing receiver solutions > use the *conventional* approach. For instance, the Lucent receiver John > just bought seems to have a discrete, independent GPS receiver (the > darker board on top), and many companies still build stand alone GPS > receivers specifically for timing application without embedding the > GPSDO logic and an OCXO. That does not seem to make much sense to me. I have ordered a Lucent RFTG-M-XO box but it has yet to arrive - however, looking at the photos published today, it does not seem clear they don't use a GPS receiver with an external clock. Not easy to tell since the guts of the Motorola module are under a shield can. There is certainly no reason to integrate a GPS receiver with the OXCO PLL and status electronics if one can buy an OEM one that takes an external clock for less money (and hastle over firmware development and licensing for the GPS side). GPSDOs are a small market compared to the overall market for OEM GPS solutions and there are economies of scale involved. And as a final note - a Datum 9390 box I have that dates to beginning of time (GPS time that is) used a Rb derived clock for the antique OEM Trimble GPS receiver board it uses so that approach has been around from the early days... -- Dave Emery N1PRE, die@dieconsulting.com DIE Consulting, Weston, Mass 02493 "An empty zombie mind with a forlorn barely readable weatherbeaten 'For Rent' sign still vainly flapping outside on the weed encrusted pole - in celebration of what could have been, but wasn't and is not to be now either."
UB
Ulrich Bangert
Wed, Dec 27, 2006 9:14 AM

Didier,

looking at your plots, i see that your 'Tau'on the main page is set to
0.5 instead of 2 as it should be with your measurement arrangement.  I
understand that due to the unity '/s' you were fooled into thinking that
you need to compute a '1/2' sample rate from your measurement
arrangemant. This is due to my mistake to mix the both terms 'sample
rate' and 'tau' into one text field. I will have to check how to name
the field best without any ambiguity. As long as this bug is there you
should simply exchange the 0.5 by 2.

After this correction your Allan plot will start with 3.06 E-9 @ 2 s.
This is VERY close to the effect of the 2 ns resolution of your counter,
so probably

The Trimble app note you pointed to says the 1 PPS output is not
quantized, so I believe that means there should not be sawtooth error.

is correct. Note however, that with 1 s dead time between measurements
we need a dead time correction that Plotter is currently not being able
to compute. The results will not be worlds apart but there will be
subtle differences.

In order to display two plots into one with the two having largely
different scales give the second plot a different vertical scale by

  1. Open chart editor

  2. In the left upper part choose the series that you want to give a new
    vertical axis

  3. In the right window open the 'General' parameter page

  4. At 'vertical axis' choose 'right' and close the chart editor to see
    the result

  5. Note that you may introduce as many new vertical axes as you want in
    the 'chart' 'axis' part of the chart editor (see the '+' sign?). Using
    the method 1/2/3 you may assign series to new created axes. Note also
    that the new axes may be shifted horizontical against their original
    position to give every axis its own gorizontal position in the plot.

I wish I had the time to write a user manual!

73 de Ulrich, DF6JB

-----Ursprüngliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Didier Juges
Gesendet: Dienstag, 26. Dezember 2006 16:40
An: Discussion of precise time and frequency measurement
Betreff: Re: [time-nuts] TIC resolution impact on GPSDO's performance

Dr Bruce Griffiths wrote:

Didier Juges wrote:

This discussion is fascinating, and as always it has prompted a
number
of other questions for me.

I understand the sawtooth correction is provided to allow

correction

of
1 PPS timing errors when the processor clock is

non-coherent with the

GPS signal (at least not intentionally), because the

processor clock is

running at a finite frequency and provides granularity to the PPS
adjustment capability.

I wanted to know if this was available in my Trimble Thunderbolt
GPSDO.

I have looked over the entire Thunderbolt manual and I

have not seen

a
mention about sawtooth correction, so I am not sure if it

is available.

Then I realized the Thunderbolt is different from most GPSDO that
have
been discussed, in that it uses the 10 MHz from the OCXO

as the clock

for the GPS receiver, so in theory, the sawtooth

correction should not

be necessary on this type of receiver since the processor

clock and the

GPS signals are coherent (maybe I should not use that

term...) when the

receiver is tracking.

Yet, the Thunderbolt spec for timing accuracy  on the 1

PPS output is

+/- 20 nS at 1 sigma, just like an ordinary GPSDO with

stand alone

receiver and what would be about a 25 MHz clock (please confirm).

The Trimble software also displays values referred to as "Timing
Output"
for the 1 PPS and the 10 MHz. The PPS value is currently

hovering around

-1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x

ppb. I am not

sure what that is based on. The manual simply refers to those as
"Estimate of UTC/GPS offset", but does not give any detail

of how they

are computed or what they mean.

The Thunderbolt User's Manual says that the 1 PPS output is the
OCXO's
10 MHz divided down. Yet, the block diagram shows the 1

PPS coming from

the CPU and support circuit, not directly coming from the

10 MHz. I

believe the 1 PPS is the 10 MHz divided down, except when

the receiver

is doing jam-sync (after power up or recovery, when the

GPS_PPS and

HW_PPS are far away from each other)

Going back to what Poul-Henning just said, I believe that for the
Thunderbolt, there actually may be three PPS signals:

GPS_PPS: what the receiver thinks the PPS should be (a theoretical
signal)
HW_PPS: the best approximation of the GPS_PPS that the

receiver can do

OCXO_PPS: the 10 MHz divided down

In jam-sync mode, the Thunderbolt forces the HW_PPS to

within 100nS

of
GPS_PPS (the closest it can get using software delays and

programmable

dividers I guess, using a 10 MHz clock) without touching

the OCXO, so in

that mode, HW_PPS and OCXO_PPS are obviously not the same.

Once the two

are within 100 nS, the Thunderbolt switches to discipline

mode and fine

tunes the OCXO to get the HW_PPS as close to GPS_PPS as

possible (20nS 1

sigma by specification.) It would appear that the receiver

should be

able to adjust its OCXO as close to GPS_PPS as the DAC will allow,
without the artificial limit of the processor clock

period. So, there

should not be a need for sawtooth correction and there

should be no

hanging bridges either.

Then, what are the "Timing Outputs"?

Does this make any sense?

Didier KO4BB


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Didier

Look at: http://trl.trimble.com/docushare/dsweb/Get/Document-8425/

There is little evidence of sawtooth corrections in graph

in the above

document, however the data were taken when SA was switched on.
To confuse matters further the document actually talks about pulse
position quantisation.

http://trl.trimble.com/docushare/dsweb/Get/Document-8424/
Shows differences between 2 Thunderbolts, again there is little
evidence
of quantisation but data may have been smoothed.

Have you any reasonably stable crystal oscillator that can

be used to

log the time interval between the Thunderbolt PPS output

and the divided

down oscillator output?
Plotting these results should quickly show evidence of any

sawtooth error.

As you say there shouldn't be any sawtooth error if the receiver
timing
is derived from the 10MHz clock which is in turned locked

to the PPS output.

However there are other sources of noise in a GPS timing

receiver which

could, in an older receiver design like this, easily be as

much as 20ns rms.

Bruce,

The Trimble app note you pointed to says the 1 PPS output is not
quantized, so I believe that means there should not be sawtooth error.

Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html

The last picture is the Thunderbolt PPS output against a free
running HP
10811 divided by 256 with two 74F161.

The plot was obtained after removing steps, outliers (few,
mostly due to
my acquisition software that puts some garbage in the output)
and drift,
using Ulrich's Plotter program.

Since I was using the HP 5334, data is taken every 2 seconds.

How can I determine from this if sawtooth error is present?

Do I need to collect data differently?

Thanks

Didier

Note: my home brew GPIB data logging software records ambient
temperature at the same rate as the TI, which I believe would
be a very
useful information. Unfortunately, I have not found yet how
to display
both curves in a meaningful way using Plotter, so I have to strip the
temperature from the dataset before feeding it to Plotter. Also,
collecting data from two instruments causes my program to sometimes
insert garbage in the output file, which causes outliers and which I
need to fix before I put this software on my web page.


time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-> bin/mailman/listinfo/time-nuts

Didier, looking at your plots, i see that your 'Tau'on the main page is set to 0.5 instead of 2 as it should be with your measurement arrangement. I understand that due to the unity '/s' you were fooled into thinking that you need to compute a '1/2' sample rate from your measurement arrangemant. This is due to my mistake to mix the both terms 'sample rate' and 'tau' into one text field. I will have to check how to name the field best without any ambiguity. As long as this bug is there you should simply exchange the 0.5 by 2. After this correction your Allan plot will start with 3.06 E-9 @ 2 s. This is VERY close to the effect of the 2 ns resolution of your counter, so probably > The Trimble app note you pointed to says the 1 PPS output is not > quantized, so I believe that means there should not be sawtooth error. is correct. Note however, that with 1 s dead time between measurements we need a dead time correction that Plotter is currently not being able to compute. The results will not be worlds apart but there will be subtle differences. In order to display two plots into one with the two having largely different scales give the second plot a different vertical scale by 1) Open chart editor 2) In the left upper part choose the series that you want to give a new vertical axis 3) In the right window open the 'General' parameter page 4) At 'vertical axis' choose 'right' and close the chart editor to see the result 5) Note that you may introduce as many new vertical axes as you want in the 'chart' 'axis' part of the chart editor (see the '+' sign?). Using the method 1/2/3 you may assign series to new created axes. Note also that the new axes may be shifted horizontical against their original position to give every axis its own gorizontal position in the plot. I wish I had the time to write a user manual! 73 de Ulrich, DF6JB > -----Ursprüngliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Didier Juges > Gesendet: Dienstag, 26. Dezember 2006 16:40 > An: Discussion of precise time and frequency measurement > Betreff: Re: [time-nuts] TIC resolution impact on GPSDO's performance > > > Dr Bruce Griffiths wrote: > > Didier Juges wrote: > > > >> This discussion is fascinating, and as always it has prompted a > >> number > >> of other questions for me. > >> > >> I understand the sawtooth correction is provided to allow > correction > >> of > >> 1 PPS timing errors when the processor clock is > non-coherent with the > >> GPS signal (at least not intentionally), because the > processor clock is > >> running at a finite frequency and provides granularity to the PPS > >> adjustment capability. > >> > >> I wanted to know if this was available in my Trimble Thunderbolt > >> GPSDO. > >> > >> I have looked over the entire Thunderbolt manual and I > have not seen > >> a > >> mention about sawtooth correction, so I am not sure if it > is available. > >> > >> Then I realized the Thunderbolt is different from most GPSDO that > >> have > >> been discussed, in that it uses the 10 MHz from the OCXO > as the clock > >> for the GPS receiver, so in theory, the sawtooth > correction should not > >> be necessary on this type of receiver since the processor > clock and the > >> GPS signals are coherent (maybe I should not use that > term...) when the > >> receiver is tracking. > >> > >> Yet, the Thunderbolt spec for timing accuracy on the 1 > PPS output is > >> +/- 20 nS at 1 sigma, just like an *ordinary* GPSDO with > stand alone > >> receiver and what would be about a 25 MHz clock (please confirm). > >> > >> The Trimble software also displays values referred to as "Timing > >> Output" > >> for the 1 PPS and the 10 MHz. The PPS value is currently > hovering around > >> -1.xx to -2.xx nS and the 10 MHz value is around +/- 0.0x > ppb. I am not > >> sure what that is based on. The manual simply refers to those as > >> "Estimate of UTC/GPS offset", but does not give any detail > of how they > >> are computed or what they mean. > >> > >> The Thunderbolt User's Manual says that the 1 PPS output is the > >> OCXO's > >> 10 MHz divided down. Yet, the block diagram shows the 1 > PPS coming from > >> the CPU and support circuit, not directly coming from the > 10 MHz. I > >> believe the 1 PPS is the 10 MHz divided down, except when > the receiver > >> is doing jam-sync (after power up or recovery, when the > GPS_PPS and > >> HW_PPS are far away from each other) > >> > >> Going back to what Poul-Henning just said, I believe that for the > >> Thunderbolt, there actually may be three PPS signals: > >> > >> GPS_PPS: what the receiver thinks the PPS should be (a theoretical > >> signal) > >> HW_PPS: the best approximation of the GPS_PPS that the > receiver can do > >> OCXO_PPS: the 10 MHz divided down > >> > >> In jam-sync mode, the Thunderbolt forces the HW_PPS to > within 100nS > >> of > >> GPS_PPS (the closest it can get using software delays and > programmable > >> dividers I guess, using a 10 MHz clock) without touching > the OCXO, so in > >> that mode, HW_PPS and OCXO_PPS are obviously not the same. > Once the two > >> are within 100 nS, the Thunderbolt switches to discipline > mode and fine > >> tunes the OCXO to get the HW_PPS as close to GPS_PPS as > possible (20nS 1 > >> sigma by specification.) It would appear that the receiver > should be > >> able to adjust its OCXO as close to GPS_PPS as the DAC will allow, > >> without the artificial limit of the processor clock > period. So, there > >> should not be a need for sawtooth correction and there > should be no > >> hanging bridges either. > >> > >> Then, what are the "Timing Outputs"? > >> > >> Does this make any sense? > >> > >> Didier KO4BB > >> > >> > >> > >> > >> _______________________________________________ > >> time-nuts mailing list > >> time-nuts@febo.com > >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> > >> > >> > > Didier > > > > Look at: http://trl.trimble.com/docushare/dsweb/Get/Document-8425/ > > > > There is little evidence of sawtooth corrections in graph > in the above > > document, however the data were taken when SA was switched on. > > To confuse matters further the document actually talks about pulse > > position quantisation. > > > > http://trl.trimble.com/docushare/dsweb/Get/Document-8424/ > > Shows differences between 2 Thunderbolts, again there is little > > evidence > > of quantisation but data may have been smoothed. > > > > Have you any reasonably stable crystal oscillator that can > be used to > > log the time interval between the Thunderbolt PPS output > and the divided > > down oscillator output? > > Plotting these results should quickly show evidence of any > sawtooth error. > > > > As you say there shouldn't be any sawtooth error if the receiver > > timing > > is derived from the 10MHz clock which is in turned locked > to the PPS output. > > However there are other sources of noise in a GPS timing > receiver which > > could, in an older receiver design like this, easily be as > much as 20ns rms. > > > > Bruce > > > > _______________________________________________ > > time-nuts mailing list > > time-nuts@febo.com > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > > > Bruce, > > The Trimble app note you pointed to says the 1 PPS output is not > quantized, so I believe that means there should not be sawtooth error. > > Look at the bottom of this page http://www.ko4bb.com/Timing/FAQ-1.html > > The last picture is the Thunderbolt PPS output against a free > running HP > 10811 divided by 256 with two 74F161. > > The plot was obtained after removing steps, outliers (few, > mostly due to > my acquisition software that puts some garbage in the output) > and drift, > using Ulrich's Plotter program. > > Since I was using the HP 5334, data is taken every 2 seconds. > > How can I determine from this if sawtooth error is present? > > Do I need to collect data differently? > > Thanks > > Didier > > Note: my home brew GPIB data logging software records ambient > temperature at the same rate as the TI, which I believe would > be a very > useful information. Unfortunately, I have not found yet how > to display > both curves in a meaningful way using Plotter, so I have to strip the > temperature from the dataset before feeding it to Plotter. Also, > collecting data from two instruments causes my program to sometimes > insert garbage in the output file, which causes outliers and which I > need to fix before I put this software on my web page. > > _______________________________________________ > time-nuts mailing list > time-nuts@febo.com > https://www.febo.com/cgi-> bin/mailman/listinfo/time-nuts >