PF
Paul Fox
Wed, Feb 6, 2013 8:59 PM
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure -- Flatness
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service manual.
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded privately.
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts. It is
Good luck.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision measurement nut
at all, but given past discussion in the list archives, i think you
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought it
worked when he last used it, but that was quite a long time ago, and i
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly wanted
it out of his house, along with a bunch of other vintage (sadly mostly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful, and
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i assume
the s/n would normally be obvious on the rear of the unit.) it bears a
sticker on the back that says "MTG PROTO 27". i think the unit came
from a place that themselves produced reference standards, and they
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there are
other clues to its age inside.
so: is this machine useful to anyone? given the error message above,
are there any measurements at all that it might still get correct, in
it's current state?
paul
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure -- Flatness
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service manual.
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
> thanks joe, and to the other couple of people who responded privately.
> i need to figure out how much of a project i want this to be, and how
> it ranks against all the other projects that i also don't have time
> for. :-)
>
> i have the full users manual (two copies!), but i don't have anything
> else that speaks GPIB. i'll play with some of the front panel
> commands you've suggested, while i think about what to do. thanks!
>
> paul
>
> j. l. trantham wrote:
> > Paul,
> >
> > First, the meter is worth something if nothing other than parts. It is
...
> > Good luck.
> >
> > Joe
> >
> > -----Original Message-----
> > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
> > Behalf Of Paul Fox
> > Sent: Monday, October 15, 2012 9:50 AM
> > To: Discussion of precise voltage measurement
> > Subject: [volt-nuts] hp 3458a advice
> >
> > hi --
> >
> > i'm not a volt-nuts regular, and not even a precision measurement nut
> > at all, but given past discussion in the list archives, i think you
> > folks might be able to help.
> >
> > i was recently given an HP 3458a multimeter. the donor thought it
> > worked when he last used it, but that was quite a long time ago, and i
> > think it was stored in an attic for some time, so i'm not too
> > surprised that it now doesn't pass self tests. :-/ he mainly wanted
> > it out of his house, along with a bunch of other vintage (sadly mostly
> > non-working) equipment he'd accumulated over the years.
> >
> > the self-test error is:
> > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
> >
> > judging from past list traffic, repairs on these units are
> > expensive, and it would still need recalibration to be useful, and
> > there are may be other parts (eeproms?) that may not be good
> > any more. i'm not really interested in pursuing any of this
> > myself.
> >
> > what's worse is that the unit doesn't have a serial number. (i assume
> > the s/n would normally be obvious on the rear of the unit.) it bears a
> > sticker on the back that says "MTG PROTO 27". i think the unit came
> > from a place that themselves produced reference standards, and they
> > may have gotten a very early model from HP to work with. that's
> > conjecture. i haven't opened the unit, so i don't know if there are
> > other clues to its age inside.
> >
> > so: is this machine useful to anyone? given the error message above,
> > are there any measurements at all that it might still get correct, in
> > it's current state?
> >
> > paul
> > =---------------------
> > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
> >
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0 degrees)
EG
Eric Garner
Wed, Feb 6, 2013 9:21 PM
Paul,
I'm interested in the 3458a. I'm located in Portland, OR since that
affects your decision. I've been lusting after a 3458a for sometime.
Since I'm on a budget most of my equipment is acquired in "needs TLC"
condition, so i'm used to a project. fortunately for me my day job has
it's yearly calibration in a few months so if I get it I can get it
cal'd properly (taking the techs out for beers goes a long way).
thanks
-eric
On Wed, Feb 6, 2013 at 12:59 PM, Paul Fox pgf@foxharp.boston.ma.us wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure -- Flatness
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service manual.
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded privately.
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts. It is
Good luck.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision measurement nut
at all, but given past discussion in the list archives, i think you
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought it
worked when he last used it, but that was quite a long time ago, and i
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly wanted
it out of his house, along with a bunch of other vintage (sadly mostly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful, and
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i assume
the s/n would normally be obvious on the rear of the unit.) it bears a
sticker on the back that says "MTG PROTO 27". i think the unit came
from a place that themselves produced reference standards, and they
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there are
other clues to its age inside.
so: is this machine useful to anyone? given the error message above,
are there any measurements at all that it might still get correct, in
it's current state?
paul
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
Paul,
I'm interested in the 3458a. I'm located in Portland, OR since that
affects your decision. I've been lusting after a 3458a for sometime.
Since I'm on a budget most of my equipment is acquired in "needs TLC"
condition, so i'm used to a project. fortunately for me my day job has
it's yearly calibration in a few months so if I get it I can get it
cal'd properly (taking the techs out for beers goes a long way).
thanks
-eric
On Wed, Feb 6, 2013 at 12:59 PM, Paul Fox <pgf@foxharp.boston.ma.us> wrote:
> resurrecting a somewhat older thread --
>
> it's now clear to me that i have little interest in (or time for)
> resurrecting my 3458a. several people on the list were interested in
> it when i first brought it up, and you were all very helpful. i'd
> rather not go through the ebay hassle, and i'd rather make it
> available to one of you folks first in any case.
>
> i've put up a small set of pictures, along with a short video of
> the startup sequence at http://www.foxharp.net/hp3458a . re-reading
> my original list message (below), i see that my startup video didn't
> capture an error message i saw back in october. so in addition to the
> "110, Calibration required -- ACAL" message in the video, i guess we
> should assume that it sometimes gives "204, "Hardware Failure -- Flatness
> DAC Convergence: 198"
>
> i haven't opened the unit, and wasn't really planning on it. if
> there's something that you'd like to know that will be clear from just
> looking, and if it might appreciably change the unit's value for you,
> let me know and i'll find a screwdriver and take some more pictures.
>
> i have two original (i believe) copies of the manual. no service manual.
> firmware version is "REV 2,1".
>
> so how about this: best offer (by private mail) under $350 gets it,
> shipping not included (it looks like shipping will be in the $40 to $80
> range, once i find a box). for obvious reasons, a local-to-boston
> transaction will be given a lot of preference.
>
> i'll wait until next wednesday evening (i.e., about a week -- feb 13)
> before choosing a lucky recipient, since i hear some people don't read
> their mail as continuously as i do.
>
> paul
>
> i wrote:
> > thanks joe, and to the other couple of people who responded privately.
> > i need to figure out how much of a project i want this to be, and how
> > it ranks against all the other projects that i also don't have time
> > for. :-)
> >
> > i have the full users manual (two copies!), but i don't have anything
> > else that speaks GPIB. i'll play with some of the front panel
> > commands you've suggested, while i think about what to do. thanks!
> >
> > paul
> >
> > j. l. trantham wrote:
> > > Paul,
> > >
> > > First, the meter is worth something if nothing other than parts. It is
> ...
> > > Good luck.
> > >
> > > Joe
> > >
> > > -----Original Message-----
> > > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
> > > Behalf Of Paul Fox
> > > Sent: Monday, October 15, 2012 9:50 AM
> > > To: Discussion of precise voltage measurement
> > > Subject: [volt-nuts] hp 3458a advice
> > >
> > > hi --
> > >
> > > i'm not a volt-nuts regular, and not even a precision measurement nut
> > > at all, but given past discussion in the list archives, i think you
> > > folks might be able to help.
> > >
> > > i was recently given an HP 3458a multimeter. the donor thought it
> > > worked when he last used it, but that was quite a long time ago, and i
> > > think it was stored in an attic for some time, so i'm not too
> > > surprised that it now doesn't pass self tests. :-/ he mainly wanted
> > > it out of his house, along with a bunch of other vintage (sadly mostly
> > > non-working) equipment he'd accumulated over the years.
> > >
> > > the self-test error is:
> > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
> > >
> > > judging from past list traffic, repairs on these units are
> > > expensive, and it would still need recalibration to be useful, and
> > > there are may be other parts (eeproms?) that may not be good
> > > any more. i'm not really interested in pursuing any of this
> > > myself.
> > >
> > > what's worse is that the unit doesn't have a serial number. (i assume
> > > the s/n would normally be obvious on the rear of the unit.) it bears a
> > > sticker on the back that says "MTG PROTO 27". i think the unit came
> > > from a place that themselves produced reference standards, and they
> > > may have gotten a very early model from HP to work with. that's
> > > conjecture. i haven't opened the unit, so i don't know if there are
> > > other clues to its age inside.
> > >
> > > so: is this machine useful to anyone? given the error message above,
> > > are there any measurements at all that it might still get correct, in
> > > it's current state?
> > >
> > > paul
> > > =---------------------
> > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
> > >
>
> =---------------------
> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0 degrees)
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
--
--Eric
_________________________________________
Eric Garner
EG
Eric Garner
Wed, Feb 6, 2013 9:22 PM
sorry all, that was meant to be off list
On Wed, Feb 6, 2013 at 1:21 PM, Eric Garner garnere@gmail.com wrote:
Paul,
I'm interested in the 3458a. I'm located in Portland, OR since that
affects your decision. I've been lusting after a 3458a for sometime.
Since I'm on a budget most of my equipment is acquired in "needs TLC"
condition, so i'm used to a project. fortunately for me my day job has
it's yearly calibration in a few months so if I get it I can get it
cal'd properly (taking the techs out for beers goes a long way).
thanks
-eric
On Wed, Feb 6, 2013 at 12:59 PM, Paul Fox pgf@foxharp.boston.ma.us wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure -- Flatness
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service manual.
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded privately.
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts. It is
Good luck.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision measurement nut
at all, but given past discussion in the list archives, i think you
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought it
worked when he last used it, but that was quite a long time ago, and i
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly wanted
it out of his house, along with a bunch of other vintage (sadly mostly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful, and
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i assume
the s/n would normally be obvious on the rear of the unit.) it bears a
sticker on the back that says "MTG PROTO 27". i think the unit came
from a place that themselves produced reference standards, and they
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there are
other clues to its age inside.
so: is this machine useful to anyone? given the error message above,
are there any measurements at all that it might still get correct, in
it's current state?
paul
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
sorry all, that was meant to be off list
On Wed, Feb 6, 2013 at 1:21 PM, Eric Garner <garnere@gmail.com> wrote:
> Paul,
>
> I'm interested in the 3458a. I'm located in Portland, OR since that
> affects your decision. I've been lusting after a 3458a for sometime.
> Since I'm on a budget most of my equipment is acquired in "needs TLC"
> condition, so i'm used to a project. fortunately for me my day job has
> it's yearly calibration in a few months so if I get it I can get it
> cal'd properly (taking the techs out for beers goes a long way).
>
> thanks
>
> -eric
>
>
>
> On Wed, Feb 6, 2013 at 12:59 PM, Paul Fox <pgf@foxharp.boston.ma.us> wrote:
>> resurrecting a somewhat older thread --
>>
>> it's now clear to me that i have little interest in (or time for)
>> resurrecting my 3458a. several people on the list were interested in
>> it when i first brought it up, and you were all very helpful. i'd
>> rather not go through the ebay hassle, and i'd rather make it
>> available to one of you folks first in any case.
>>
>> i've put up a small set of pictures, along with a short video of
>> the startup sequence at http://www.foxharp.net/hp3458a . re-reading
>> my original list message (below), i see that my startup video didn't
>> capture an error message i saw back in october. so in addition to the
>> "110, Calibration required -- ACAL" message in the video, i guess we
>> should assume that it sometimes gives "204, "Hardware Failure -- Flatness
>> DAC Convergence: 198"
>>
>> i haven't opened the unit, and wasn't really planning on it. if
>> there's something that you'd like to know that will be clear from just
>> looking, and if it might appreciably change the unit's value for you,
>> let me know and i'll find a screwdriver and take some more pictures.
>>
>> i have two original (i believe) copies of the manual. no service manual.
>> firmware version is "REV 2,1".
>>
>> so how about this: best offer (by private mail) under $350 gets it,
>> shipping not included (it looks like shipping will be in the $40 to $80
>> range, once i find a box). for obvious reasons, a local-to-boston
>> transaction will be given a lot of preference.
>>
>> i'll wait until next wednesday evening (i.e., about a week -- feb 13)
>> before choosing a lucky recipient, since i hear some people don't read
>> their mail as continuously as i do.
>>
>> paul
>>
>> i wrote:
>> > thanks joe, and to the other couple of people who responded privately.
>> > i need to figure out how much of a project i want this to be, and how
>> > it ranks against all the other projects that i also don't have time
>> > for. :-)
>> >
>> > i have the full users manual (two copies!), but i don't have anything
>> > else that speaks GPIB. i'll play with some of the front panel
>> > commands you've suggested, while i think about what to do. thanks!
>> >
>> > paul
>> >
>> > j. l. trantham wrote:
>> > > Paul,
>> > >
>> > > First, the meter is worth something if nothing other than parts. It is
>> ...
>> > > Good luck.
>> > >
>> > > Joe
>> > >
>> > > -----Original Message-----
>> > > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
>> > > Behalf Of Paul Fox
>> > > Sent: Monday, October 15, 2012 9:50 AM
>> > > To: Discussion of precise voltage measurement
>> > > Subject: [volt-nuts] hp 3458a advice
>> > >
>> > > hi --
>> > >
>> > > i'm not a volt-nuts regular, and not even a precision measurement nut
>> > > at all, but given past discussion in the list archives, i think you
>> > > folks might be able to help.
>> > >
>> > > i was recently given an HP 3458a multimeter. the donor thought it
>> > > worked when he last used it, but that was quite a long time ago, and i
>> > > think it was stored in an attic for some time, so i'm not too
>> > > surprised that it now doesn't pass self tests. :-/ he mainly wanted
>> > > it out of his house, along with a bunch of other vintage (sadly mostly
>> > > non-working) equipment he'd accumulated over the years.
>> > >
>> > > the self-test error is:
>> > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
>> > >
>> > > judging from past list traffic, repairs on these units are
>> > > expensive, and it would still need recalibration to be useful, and
>> > > there are may be other parts (eeproms?) that may not be good
>> > > any more. i'm not really interested in pursuing any of this
>> > > myself.
>> > >
>> > > what's worse is that the unit doesn't have a serial number. (i assume
>> > > the s/n would normally be obvious on the rear of the unit.) it bears a
>> > > sticker on the back that says "MTG PROTO 27". i think the unit came
>> > > from a place that themselves produced reference standards, and they
>> > > may have gotten a very early model from HP to work with. that's
>> > > conjecture. i haven't opened the unit, so i don't know if there are
>> > > other clues to its age inside.
>> > >
>> > > so: is this machine useful to anyone? given the error message above,
>> > > are there any measurements at all that it might still get correct, in
>> > > it's current state?
>> > >
>> > > paul
>> > > =---------------------
>> > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
>> > >
>>
>> =---------------------
>> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0 degrees)
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
>
>
> --
> --Eric
> _________________________________________
> Eric Garner
--
--Eric
_________________________________________
Eric Garner
PF
Paul Fox
Thu, Feb 7, 2013 2:53 PM
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure -- Flatness
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service manual.
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded privately.
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts. It is
Good luck.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision measurement nut
at all, but given past discussion in the list archives, i think you
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought it
worked when he last used it, but that was quite a long time ago, and i
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly wanted
it out of his house, along with a bunch of other vintage (sadly mostly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful, and
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i assume
the s/n would normally be obvious on the rear of the unit.) it bears a
sticker on the back that says "MTG PROTO 27". i think the unit came
from a place that themselves produced reference standards, and they
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there are
other clues to its age inside.
so: is this machine useful to anyone? given the error message above,
are there any measurements at all that it might still get correct, in
it's current state?
paul
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
> resurrecting a somewhat older thread --
>
> it's now clear to me that i have little interest in (or time for)
> resurrecting my 3458a. several people on the list were interested in
> it when i first brought it up, and you were all very helpful. i'd
> rather not go through the ebay hassle, and i'd rather make it
> available to one of you folks first in any case.
>
> i've put up a small set of pictures, along with a short video of
> the startup sequence at http://www.foxharp.net/hp3458a . re-reading
> my original list message (below), i see that my startup video didn't
> capture an error message i saw back in october. so in addition to the
> "110, Calibration required -- ACAL" message in the video, i guess we
> should assume that it sometimes gives "204, "Hardware Failure -- Flatness
> DAC Convergence: 198"
>
> i haven't opened the unit, and wasn't really planning on it. if
> there's something that you'd like to know that will be clear from just
> looking, and if it might appreciably change the unit's value for you,
> let me know and i'll find a screwdriver and take some more pictures.
>
> i have two original (i believe) copies of the manual. no service manual.
> firmware version is "REV 2,1".
>
> so how about this: best offer (by private mail) under $350 gets it,
> shipping not included (it looks like shipping will be in the $40 to $80
> range, once i find a box). for obvious reasons, a local-to-boston
> transaction will be given a lot of preference.
>
> i'll wait until next wednesday evening (i.e., about a week -- feb 13)
> before choosing a lucky recipient, since i hear some people don't read
> their mail as continuously as i do.
>
> paul
>
> i wrote:
> > thanks joe, and to the other couple of people who responded privately.
> > i need to figure out how much of a project i want this to be, and how
> > it ranks against all the other projects that i also don't have time
> > for. :-)
> >
> > i have the full users manual (two copies!), but i don't have anything
> > else that speaks GPIB. i'll play with some of the front panel
> > commands you've suggested, while i think about what to do. thanks!
> >
> > paul
> >
> > j. l. trantham wrote:
> > > Paul,
> > >
> > > First, the meter is worth something if nothing other than parts. It is
> ...
> > > Good luck.
> > >
> > > Joe
> > >
> > > -----Original Message-----
> > > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
> > > Behalf Of Paul Fox
> > > Sent: Monday, October 15, 2012 9:50 AM
> > > To: Discussion of precise voltage measurement
> > > Subject: [volt-nuts] hp 3458a advice
> > >
> > > hi --
> > >
> > > i'm not a volt-nuts regular, and not even a precision measurement nut
> > > at all, but given past discussion in the list archives, i think you
> > > folks might be able to help.
> > >
> > > i was recently given an HP 3458a multimeter. the donor thought it
> > > worked when he last used it, but that was quite a long time ago, and i
> > > think it was stored in an attic for some time, so i'm not too
> > > surprised that it now doesn't pass self tests. :-/ he mainly wanted
> > > it out of his house, along with a bunch of other vintage (sadly mostly
> > > non-working) equipment he'd accumulated over the years.
> > >
> > > the self-test error is:
> > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
> > >
> > > judging from past list traffic, repairs on these units are
> > > expensive, and it would still need recalibration to be useful, and
> > > there are may be other parts (eeproms?) that may not be good
> > > any more. i'm not really interested in pursuing any of this
> > > myself.
> > >
> > > what's worse is that the unit doesn't have a serial number. (i assume
> > > the s/n would normally be obvious on the rear of the unit.) it bears a
> > > sticker on the back that says "MTG PROTO 27". i think the unit came
> > > from a place that themselves produced reference standards, and they
> > > may have gotten a very early model from HP to work with. that's
> > > conjecture. i haven't opened the unit, so i don't know if there are
> > > other clues to its age inside.
> > >
> > > so: is this machine useful to anyone? given the error message above,
> > > are there any measurements at all that it might still get correct, in
> > > it's current state?
> > >
> > > paul
> > > =---------------------
> > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
> > >
>
> =---------------------
> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0 degrees)
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 19.2 degrees)
JL
J. L. Trantham
Fri, Feb 8, 2013 1:59 AM
Paul,
The 'value' of a 3458A is based on several issues, not the least of which is
supply and demand. You only have to look at theBay to get an idea of what
units are selling for, both working and non-working. The fact that the unit
passes it's 'Self Test' is not a guarantee that it would be able to be
calibrated to published performance specs.
Tha Agilent website is very useful in determining an 'upper bound' on the
value. The 3458A is currently orderable and currently supported. A new
3458A (without options) is available at $9,085 and would come with a fresh
calibration and warranty.
Repair Per Incident is available for $2,660.64. This means that if you have
a complete unit, they will repair it for that price. That also comes with a
fresh calibration. I don't know if that price includes all recommended
upgrades as listed in the Service Notes, of which there are 20, 6 of which
list 'Modification Recommended' and one of which lists 'Modification
Available'. Perhaps someone on the forum knows the answer to this.
Certainly Gary Biermann at Agilent can answer this.
The fact that your unit does not have a serial number sticker but instead
'MTG PROTO 27' would be concerning as to whether Agilent would consider this
a 'complete 3458A', and thus a candidate for repair, or, rather, a
'prototype' or some other descriptor that would imply that it looks like a
3458A but is really not a 3458A.
Both new and newly calibrated units are candidates to add a Repair
Agreement, available for $178.68 per year with up to 5 year blocks available
for purchase.
I note you mentioned firmware REV 2,1. The current firmware is 9,1.
Therefore it would appear that you have a rather old and probably not
'updated' unit. That plus the 'stored in an attic' history would be
concerning about the ability to repair the unit without changing several
assemblies, etc.
Assuming you could get the unit repaired and updated by Agilent, complete
with a fresh calibration, then the maximum value would be $9,085 minus
$2,660.64 minus what ever value you place on 'new' over 'old'.
Then you are left with an older unit, made in the USA, with no (or perhaps a
reduced) warranty as opposed to a new unit, made in Malaysia, with a
warranty. The question of whether a 'new' Malaysian unit or an 'old'
American unit is better is the subject of further discussion and
speculation.
If the unit cannot be repaired by Agilent, then, in my opinion, the value
goes down and approaches the value of the hardware, transformer, etc., with
serious questions about the state and reparability of the assemblies,
particularly given the 'stored in an attic' history.
Gary Biermann would be very helpful in getting the history and reparability
of the unit. He would be worth contacting about these issues. He works at
the Loveland Calibration Facility.
Hope this helps.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Thursday, February 07, 2013 8:54 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure -- Flatness
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service manual.
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded privately.
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts.
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision measurement
at all, but given past discussion in the list archives, i think you
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought it
worked when he last used it, but that was quite a long time ago,
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly
it out of his house, along with a bunch of other vintage (sadly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful, and
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i
the s/n would normally be obvious on the rear of the unit.) it
sticker on the back that says "MTG PROTO 27". i think the unit
from a place that themselves produced reference standards, and they
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there
other clues to its age inside.
so: is this machine useful to anyone? given the error message
are there any measurements at all that it might still get correct,
and follow the instructions there.
Paul,
The 'value' of a 3458A is based on several issues, not the least of which is
supply and demand. You only have to look at theBay to get an idea of what
units are selling for, both working and non-working. The fact that the unit
passes it's 'Self Test' is not a guarantee that it would be able to be
calibrated to published performance specs.
Tha Agilent website is very useful in determining an 'upper bound' on the
value. The 3458A is currently orderable and currently supported. A new
3458A (without options) is available at $9,085 and would come with a fresh
calibration and warranty.
Repair Per Incident is available for $2,660.64. This means that if you have
a complete unit, they will repair it for that price. That also comes with a
fresh calibration. I don't know if that price includes all recommended
upgrades as listed in the Service Notes, of which there are 20, 6 of which
list 'Modification Recommended' and one of which lists 'Modification
Available'. Perhaps someone on the forum knows the answer to this.
Certainly Gary Biermann at Agilent can answer this.
The fact that your unit does not have a serial number sticker but instead
'MTG PROTO 27' would be concerning as to whether Agilent would consider this
a 'complete 3458A', and thus a candidate for repair, or, rather, a
'prototype' or some other descriptor that would imply that it looks like a
3458A but is really not a 3458A.
Both new and newly calibrated units are candidates to add a Repair
Agreement, available for $178.68 per year with up to 5 year blocks available
for purchase.
I note you mentioned firmware REV 2,1. The current firmware is 9,1.
Therefore it would appear that you have a rather old and probably not
'updated' unit. That plus the 'stored in an attic' history would be
concerning about the ability to repair the unit without changing several
assemblies, etc.
Assuming you could get the unit repaired and updated by Agilent, complete
with a fresh calibration, then the maximum value would be $9,085 minus
$2,660.64 minus what ever value you place on 'new' over 'old'.
Then you are left with an older unit, made in the USA, with no (or perhaps a
reduced) warranty as opposed to a new unit, made in Malaysia, with a
warranty. The question of whether a 'new' Malaysian unit or an 'old'
American unit is better is the subject of further discussion and
speculation.
If the unit cannot be repaired by Agilent, then, in my opinion, the value
goes down and approaches the value of the hardware, transformer, etc., with
serious questions about the state and reparability of the assemblies,
particularly given the 'stored in an attic' history.
Gary Biermann would be very helpful in getting the history and reparability
of the unit. He would be worth contacting about these issues. He works at
the Loveland Calibration Facility.
Hope this helps.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Thursday, February 07, 2013 8:54 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
> resurrecting a somewhat older thread --
>
> it's now clear to me that i have little interest in (or time for)
> resurrecting my 3458a. several people on the list were interested in
> it when i first brought it up, and you were all very helpful. i'd
> rather not go through the ebay hassle, and i'd rather make it
> available to one of you folks first in any case.
>
> i've put up a small set of pictures, along with a short video of
> the startup sequence at http://www.foxharp.net/hp3458a . re-reading
> my original list message (below), i see that my startup video didn't
> capture an error message i saw back in october. so in addition to the
> "110, Calibration required -- ACAL" message in the video, i guess we
> should assume that it sometimes gives "204, "Hardware Failure -- Flatness
> DAC Convergence: 198"
>
> i haven't opened the unit, and wasn't really planning on it. if
> there's something that you'd like to know that will be clear from just
> looking, and if it might appreciably change the unit's value for you,
> let me know and i'll find a screwdriver and take some more pictures.
>
> i have two original (i believe) copies of the manual. no service manual.
> firmware version is "REV 2,1".
>
> so how about this: best offer (by private mail) under $350 gets it,
> shipping not included (it looks like shipping will be in the $40 to $80
> range, once i find a box). for obvious reasons, a local-to-boston
> transaction will be given a lot of preference.
>
> i'll wait until next wednesday evening (i.e., about a week -- feb 13)
> before choosing a lucky recipient, since i hear some people don't read
> their mail as continuously as i do.
>
> paul
>
> i wrote:
> > thanks joe, and to the other couple of people who responded privately.
> > i need to figure out how much of a project i want this to be, and how
> > it ranks against all the other projects that i also don't have time
> > for. :-)
> >
> > i have the full users manual (two copies!), but i don't have anything
> > else that speaks GPIB. i'll play with some of the front panel
> > commands you've suggested, while i think about what to do. thanks!
> >
> > paul
> >
> > j. l. trantham wrote:
> > > Paul,
> > >
> > > First, the meter is worth something if nothing other than parts.
It is
> ...
> > > Good luck.
> > >
> > > Joe
> > >
> > > -----Original Message-----
> > > From: volt-nuts-bounces@febo.com
[mailto:volt-nuts-bounces@febo.com] On
> > > Behalf Of Paul Fox
> > > Sent: Monday, October 15, 2012 9:50 AM
> > > To: Discussion of precise voltage measurement
> > > Subject: [volt-nuts] hp 3458a advice
> > >
> > > hi --
> > >
> > > i'm not a volt-nuts regular, and not even a precision measurement
nut
> > > at all, but given past discussion in the list archives, i think you
> > > folks might be able to help.
> > >
> > > i was recently given an HP 3458a multimeter. the donor thought it
> > > worked when he last used it, but that was quite a long time ago,
and i
> > > think it was stored in an attic for some time, so i'm not too
> > > surprised that it now doesn't pass self tests. :-/ he mainly
wanted
> > > it out of his house, along with a bunch of other vintage (sadly
mostly
> > > non-working) equipment he'd accumulated over the years.
> > >
> > > the self-test error is:
> > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
> > >
> > > judging from past list traffic, repairs on these units are
> > > expensive, and it would still need recalibration to be useful, and
> > > there are may be other parts (eeproms?) that may not be good
> > > any more. i'm not really interested in pursuing any of this
> > > myself.
> > >
> > > what's worse is that the unit doesn't have a serial number. (i
assume
> > > the s/n would normally be obvious on the rear of the unit.) it
bears a
> > > sticker on the back that says "MTG PROTO 27". i think the unit
came
> > > from a place that themselves produced reference standards, and they
> > > may have gotten a very early model from HP to work with. that's
> > > conjecture. i haven't opened the unit, so i don't know if there
are
> > > other clues to its age inside.
> > >
> > > so: is this machine useful to anyone? given the error message
above,
> > > are there any measurements at all that it might still get correct,
in
> > > it's current state?
> > >
> > > paul
> > > =---------------------
> > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 63.1
degrees)
> > >
>
> =---------------------
> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0
degrees)
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 19.2 degrees)
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
JP
John Phillips
Fri, Feb 8, 2013 2:34 AM
You should be able to any unit repaired by Agilent as long as all the parts
are there for the $3000 flat rate repair and calibration charge.
On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham jltran@att.net wrote:
Paul,
The 'value' of a 3458A is based on several issues, not the least of which
is
supply and demand. You only have to look at theBay to get an idea of what
units are selling for, both working and non-working. The fact that the
unit
passes it's 'Self Test' is not a guarantee that it would be able to be
calibrated to published performance specs.
Tha Agilent website is very useful in determining an 'upper bound' on the
value. The 3458A is currently orderable and currently supported. A new
3458A (without options) is available at $9,085 and would come with a fresh
calibration and warranty.
Repair Per Incident is available for $2,660.64. This means that if you
have
a complete unit, they will repair it for that price. That also comes with
a
fresh calibration. I don't know if that price includes all recommended
upgrades as listed in the Service Notes, of which there are 20, 6 of which
list 'Modification Recommended' and one of which lists 'Modification
Available'. Perhaps someone on the forum knows the answer to this.
Certainly Gary Biermann at Agilent can answer this.
The fact that your unit does not have a serial number sticker but instead
'MTG PROTO 27' would be concerning as to whether Agilent would consider
this
a 'complete 3458A', and thus a candidate for repair, or, rather, a
'prototype' or some other descriptor that would imply that it looks like a
3458A but is really not a 3458A.
Both new and newly calibrated units are candidates to add a Repair
Agreement, available for $178.68 per year with up to 5 year blocks
available
for purchase.
I note you mentioned firmware REV 2,1. The current firmware is 9,1.
Therefore it would appear that you have a rather old and probably not
'updated' unit. That plus the 'stored in an attic' history would be
concerning about the ability to repair the unit without changing several
assemblies, etc.
Assuming you could get the unit repaired and updated by Agilent, complete
with a fresh calibration, then the maximum value would be $9,085 minus
$2,660.64 minus what ever value you place on 'new' over 'old'.
Then you are left with an older unit, made in the USA, with no (or perhaps
a
reduced) warranty as opposed to a new unit, made in Malaysia, with a
warranty. The question of whether a 'new' Malaysian unit or an 'old'
American unit is better is the subject of further discussion and
speculation.
If the unit cannot be repaired by Agilent, then, in my opinion, the value
goes down and approaches the value of the hardware, transformer, etc., with
serious questions about the state and reparability of the assemblies,
particularly given the 'stored in an attic' history.
Gary Biermann would be very helpful in getting the history and reparability
of the unit. He would be worth contacting about these issues. He works at
the Loveland Calibration Facility.
Hope this helps.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Thursday, February 07, 2013 8:54 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure --
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to $80
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts.
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision measurement
at all, but given past discussion in the list archives, i think
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought it
worked when he last used it, but that was quite a long time ago,
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly
it out of his house, along with a bunch of other vintage (sadly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful, and
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i
the s/n would normally be obvious on the rear of the unit.) it
sticker on the back that says "MTG PROTO 27". i think the unit
from a place that themselves produced reference standards, and
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there
other clues to its age inside.
so: is this machine useful to anyone? given the error message
are there any measurements at all that it might still get correct,
and follow the instructions there.
You should be able to any unit repaired by Agilent as long as all the parts
are there for the $3000 flat rate repair and calibration charge.
On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham <jltran@att.net> wrote:
> Paul,
>
> The 'value' of a 3458A is based on several issues, not the least of which
> is
> supply and demand. You only have to look at theBay to get an idea of what
> units are selling for, both working and non-working. The fact that the
> unit
> passes it's 'Self Test' is not a guarantee that it would be able to be
> calibrated to published performance specs.
>
> Tha Agilent website is very useful in determining an 'upper bound' on the
> value. The 3458A is currently orderable and currently supported. A new
> 3458A (without options) is available at $9,085 and would come with a fresh
> calibration and warranty.
>
> Repair Per Incident is available for $2,660.64. This means that if you
> have
> a complete unit, they will repair it for that price. That also comes with
> a
> fresh calibration. I don't know if that price includes all recommended
> upgrades as listed in the Service Notes, of which there are 20, 6 of which
> list 'Modification Recommended' and one of which lists 'Modification
> Available'. Perhaps someone on the forum knows the answer to this.
> Certainly Gary Biermann at Agilent can answer this.
>
> The fact that your unit does not have a serial number sticker but instead
> 'MTG PROTO 27' would be concerning as to whether Agilent would consider
> this
> a 'complete 3458A', and thus a candidate for repair, or, rather, a
> 'prototype' or some other descriptor that would imply that it looks like a
> 3458A but is really not a 3458A.
>
> Both new and newly calibrated units are candidates to add a Repair
> Agreement, available for $178.68 per year with up to 5 year blocks
> available
> for purchase.
>
> I note you mentioned firmware REV 2,1. The current firmware is 9,1.
> Therefore it would appear that you have a rather old and probably not
> 'updated' unit. That plus the 'stored in an attic' history would be
> concerning about the ability to repair the unit without changing several
> assemblies, etc.
>
> Assuming you could get the unit repaired and updated by Agilent, complete
> with a fresh calibration, then the maximum value would be $9,085 minus
> $2,660.64 minus what ever value you place on 'new' over 'old'.
>
> Then you are left with an older unit, made in the USA, with no (or perhaps
> a
> reduced) warranty as opposed to a new unit, made in Malaysia, with a
> warranty. The question of whether a 'new' Malaysian unit or an 'old'
> American unit is better is the subject of further discussion and
> speculation.
>
> If the unit cannot be repaired by Agilent, then, in my opinion, the value
> goes down and approaches the value of the hardware, transformer, etc., with
> serious questions about the state and reparability of the assemblies,
> particularly given the 'stored in an attic' history.
>
> Gary Biermann would be very helpful in getting the history and reparability
> of the unit. He would be worth contacting about these issues. He works at
> the Loveland Calibration Facility.
>
> Hope this helps.
>
> Joe
>
> -----Original Message-----
> From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
> Behalf Of Paul Fox
> Sent: Thursday, February 07, 2013 8:54 AM
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>
> part of me hates to do this, but... there have been some rule-breakers
> that ignored the "under $350" limit, and now they've ruined it for
> everyone. ;-)
>
> clearly the research i did into prices for "HP3458a, needs repair" was
> woefully inadequate.
>
> i guess the easiest thing to do is to simply remove the "under $350"
> part. that way everyone's working under the same assumptions. again,
> sorry. local buyers still get preference. (and, if you'd like to chew
> me out for being a bozo, that's fine, but please do it privately. maybe
> i should have used ebay in the first place. lesson learned.)
>
> paul
>
> i wrote:
> > resurrecting a somewhat older thread --
> >
> > it's now clear to me that i have little interest in (or time for)
> > resurrecting my 3458a. several people on the list were interested in
> > it when i first brought it up, and you were all very helpful. i'd
> > rather not go through the ebay hassle, and i'd rather make it
> > available to one of you folks first in any case.
> >
> > i've put up a small set of pictures, along with a short video of
> > the startup sequence at http://www.foxharp.net/hp3458a . re-reading
> > my original list message (below), i see that my startup video didn't
> > capture an error message i saw back in october. so in addition to the
> > "110, Calibration required -- ACAL" message in the video, i guess we
> > should assume that it sometimes gives "204, "Hardware Failure --
> Flatness
> > DAC Convergence: 198"
> >
> > i haven't opened the unit, and wasn't really planning on it. if
> > there's something that you'd like to know that will be clear from just
> > looking, and if it might appreciably change the unit's value for you,
> > let me know and i'll find a screwdriver and take some more pictures.
> >
> > i have two original (i believe) copies of the manual. no service
> manual.
> > firmware version is "REV 2,1".
> >
> > so how about this: best offer (by private mail) under $350 gets it,
> > shipping not included (it looks like shipping will be in the $40 to $80
> > range, once i find a box). for obvious reasons, a local-to-boston
> > transaction will be given a lot of preference.
> >
> > i'll wait until next wednesday evening (i.e., about a week -- feb 13)
> > before choosing a lucky recipient, since i hear some people don't read
> > their mail as continuously as i do.
> >
> > paul
> >
> > i wrote:
> > > thanks joe, and to the other couple of people who responded
> privately.
>
> > > i need to figure out how much of a project i want this to be, and how
> > > it ranks against all the other projects that i also don't have time
> > > for. :-)
> > >
> > > i have the full users manual (two copies!), but i don't have anything
> > > else that speaks GPIB. i'll play with some of the front panel
> > > commands you've suggested, while i think about what to do. thanks!
> > >
> > > paul
> > >
> > > j. l. trantham wrote:
> > > > Paul,
> > > >
> > > > First, the meter is worth something if nothing other than parts.
> It is
> > ...
> > > > Good luck.
> > > >
> > > > Joe
> > > >
> > > > -----Original Message-----
> > > > From: volt-nuts-bounces@febo.com
> [mailto:volt-nuts-bounces@febo.com] On
> > > > Behalf Of Paul Fox
> > > > Sent: Monday, October 15, 2012 9:50 AM
> > > > To: Discussion of precise voltage measurement
> > > > Subject: [volt-nuts] hp 3458a advice
> > > >
> > > > hi --
> > > >
> > > > i'm not a volt-nuts regular, and not even a precision measurement
> nut
> > > > at all, but given past discussion in the list archives, i think
> you
> > > > folks might be able to help.
> > > >
> > > > i was recently given an HP 3458a multimeter. the donor thought it
> > > > worked when he last used it, but that was quite a long time ago,
> and i
> > > > think it was stored in an attic for some time, so i'm not too
> > > > surprised that it now doesn't pass self tests. :-/ he mainly
> wanted
> > > > it out of his house, along with a bunch of other vintage (sadly
> mostly
> > > > non-working) equipment he'd accumulated over the years.
> > > >
> > > > the self-test error is:
> > > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
> 198"
> > > >
> > > > judging from past list traffic, repairs on these units are
> > > > expensive, and it would still need recalibration to be useful, and
> > > > there are may be other parts (eeproms?) that may not be good
> > > > any more. i'm not really interested in pursuing any of this
> > > > myself.
> > > >
> > > > what's worse is that the unit doesn't have a serial number. (i
> assume
> > > > the s/n would normally be obvious on the rear of the unit.) it
> bears a
> > > > sticker on the back that says "MTG PROTO 27". i think the unit
> came
> > > > from a place that themselves produced reference standards, and
> they
> > > > may have gotten a very early model from HP to work with. that's
> > > > conjecture. i haven't opened the unit, so i don't know if there
> are
> > > > other clues to its age inside.
> > > >
> > > > so: is this machine useful to anyone? given the error message
> above,
> > > > are there any measurements at all that it might still get correct,
> in
> > > > it's current state?
> > > >
> > > > paul
> > > > =---------------------
> > > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's
> 63.1
> degrees)
> > > >
> >
> > =---------------------
> > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0
> degrees)
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> =---------------------
> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 19.2
> degrees)
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
--
John Phillips
TM
Tom Miller
Fri, Feb 8, 2013 3:12 AM
They might not take it in without a serial number.
Regards
----- Original Message -----
From: "John Phillips" john.phillips0@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Thursday, February 07, 2013 9:34 PM
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
You should be able to any unit repaired by Agilent as long as all the
parts
are there for the $3000 flat rate repair and calibration charge.
On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham jltran@att.net wrote:
Paul,
The 'value' of a 3458A is based on several issues, not the least of which
is
supply and demand. You only have to look at theBay to get an idea of
what
units are selling for, both working and non-working. The fact that the
unit
passes it's 'Self Test' is not a guarantee that it would be able to be
calibrated to published performance specs.
Tha Agilent website is very useful in determining an 'upper bound' on the
value. The 3458A is currently orderable and currently supported. A new
3458A (without options) is available at $9,085 and would come with a
fresh
calibration and warranty.
Repair Per Incident is available for $2,660.64. This means that if you
have
a complete unit, they will repair it for that price. That also comes
with
a
fresh calibration. I don't know if that price includes all recommended
upgrades as listed in the Service Notes, of which there are 20, 6 of
which
list 'Modification Recommended' and one of which lists 'Modification
Available'. Perhaps someone on the forum knows the answer to this.
Certainly Gary Biermann at Agilent can answer this.
The fact that your unit does not have a serial number sticker but instead
'MTG PROTO 27' would be concerning as to whether Agilent would consider
this
a 'complete 3458A', and thus a candidate for repair, or, rather, a
'prototype' or some other descriptor that would imply that it looks like
a
3458A but is really not a 3458A.
Both new and newly calibrated units are candidates to add a Repair
Agreement, available for $178.68 per year with up to 5 year blocks
available
for purchase.
I note you mentioned firmware REV 2,1. The current firmware is 9,1.
Therefore it would appear that you have a rather old and probably not
'updated' unit. That plus the 'stored in an attic' history would be
concerning about the ability to repair the unit without changing several
assemblies, etc.
Assuming you could get the unit repaired and updated by Agilent, complete
with a fresh calibration, then the maximum value would be $9,085 minus
$2,660.64 minus what ever value you place on 'new' over 'old'.
Then you are left with an older unit, made in the USA, with no (or
perhaps
a
reduced) warranty as opposed to a new unit, made in Malaysia, with a
warranty. The question of whether a 'new' Malaysian unit or an 'old'
American unit is better is the subject of further discussion and
speculation.
If the unit cannot be repaired by Agilent, then, in my opinion, the value
goes down and approaches the value of the hardware, transformer, etc.,
with
serious questions about the state and reparability of the assemblies,
particularly given the 'stored in an attic' history.
Gary Biermann would be very helpful in getting the history and
reparability
of the unit. He would be worth contacting about these issues. He works
at
the Loveland Calibration Facility.
Hope this helps.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Paul Fox
Sent: Thursday, February 07, 2013 8:54 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure --
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded
i need to figure out how much of a project i want this to be, and
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts.
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision
at all, but given past discussion in the list archives, i think
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought
worked when he last used it, but that was quite a long time ago,
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly
it out of his house, along with a bunch of other vintage (sadly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful,
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i
the s/n would normally be obvious on the rear of the unit.) it
sticker on the back that says "MTG PROTO 27". i think the unit
from a place that themselves produced reference standards, and
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there
other clues to its age inside.
so: is this machine useful to anyone? given the error message
are there any measurements at all that it might still get
and follow the instructions there.
They might not take it in without a serial number.
Regards
----- Original Message -----
From: "John Phillips" <john.phillips0@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Thursday, February 07, 2013 9:34 PM
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
> You should be able to any unit repaired by Agilent as long as all the
> parts
> are there for the $3000 flat rate repair and calibration charge.
>
> On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham <jltran@att.net> wrote:
>
>> Paul,
>>
>> The 'value' of a 3458A is based on several issues, not the least of which
>> is
>> supply and demand. You only have to look at theBay to get an idea of
>> what
>> units are selling for, both working and non-working. The fact that the
>> unit
>> passes it's 'Self Test' is not a guarantee that it would be able to be
>> calibrated to published performance specs.
>>
>> Tha Agilent website is very useful in determining an 'upper bound' on the
>> value. The 3458A is currently orderable and currently supported. A new
>> 3458A (without options) is available at $9,085 and would come with a
>> fresh
>> calibration and warranty.
>>
>> Repair Per Incident is available for $2,660.64. This means that if you
>> have
>> a complete unit, they will repair it for that price. That also comes
>> with
>> a
>> fresh calibration. I don't know if that price includes all recommended
>> upgrades as listed in the Service Notes, of which there are 20, 6 of
>> which
>> list 'Modification Recommended' and one of which lists 'Modification
>> Available'. Perhaps someone on the forum knows the answer to this.
>> Certainly Gary Biermann at Agilent can answer this.
>>
>> The fact that your unit does not have a serial number sticker but instead
>> 'MTG PROTO 27' would be concerning as to whether Agilent would consider
>> this
>> a 'complete 3458A', and thus a candidate for repair, or, rather, a
>> 'prototype' or some other descriptor that would imply that it looks like
>> a
>> 3458A but is really not a 3458A.
>>
>> Both new and newly calibrated units are candidates to add a Repair
>> Agreement, available for $178.68 per year with up to 5 year blocks
>> available
>> for purchase.
>>
>> I note you mentioned firmware REV 2,1. The current firmware is 9,1.
>> Therefore it would appear that you have a rather old and probably not
>> 'updated' unit. That plus the 'stored in an attic' history would be
>> concerning about the ability to repair the unit without changing several
>> assemblies, etc.
>>
>> Assuming you could get the unit repaired and updated by Agilent, complete
>> with a fresh calibration, then the maximum value would be $9,085 minus
>> $2,660.64 minus what ever value you place on 'new' over 'old'.
>>
>> Then you are left with an older unit, made in the USA, with no (or
>> perhaps
>> a
>> reduced) warranty as opposed to a new unit, made in Malaysia, with a
>> warranty. The question of whether a 'new' Malaysian unit or an 'old'
>> American unit is better is the subject of further discussion and
>> speculation.
>>
>> If the unit cannot be repaired by Agilent, then, in my opinion, the value
>> goes down and approaches the value of the hardware, transformer, etc.,
>> with
>> serious questions about the state and reparability of the assemblies,
>> particularly given the 'stored in an attic' history.
>>
>> Gary Biermann would be very helpful in getting the history and
>> reparability
>> of the unit. He would be worth contacting about these issues. He works
>> at
>> the Loveland Calibration Facility.
>>
>> Hope this helps.
>>
>> Joe
>>
>> -----Original Message-----
>> From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
>> Behalf Of Paul Fox
>> Sent: Thursday, February 07, 2013 8:54 AM
>> To: Discussion of precise voltage measurement
>> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>>
>> part of me hates to do this, but... there have been some rule-breakers
>> that ignored the "under $350" limit, and now they've ruined it for
>> everyone. ;-)
>>
>> clearly the research i did into prices for "HP3458a, needs repair" was
>> woefully inadequate.
>>
>> i guess the easiest thing to do is to simply remove the "under $350"
>> part. that way everyone's working under the same assumptions. again,
>> sorry. local buyers still get preference. (and, if you'd like to chew
>> me out for being a bozo, that's fine, but please do it privately. maybe
>> i should have used ebay in the first place. lesson learned.)
>>
>> paul
>>
>> i wrote:
>> > resurrecting a somewhat older thread --
>> >
>> > it's now clear to me that i have little interest in (or time for)
>> > resurrecting my 3458a. several people on the list were interested in
>> > it when i first brought it up, and you were all very helpful. i'd
>> > rather not go through the ebay hassle, and i'd rather make it
>> > available to one of you folks first in any case.
>> >
>> > i've put up a small set of pictures, along with a short video of
>> > the startup sequence at http://www.foxharp.net/hp3458a . re-reading
>> > my original list message (below), i see that my startup video didn't
>> > capture an error message i saw back in october. so in addition to the
>> > "110, Calibration required -- ACAL" message in the video, i guess we
>> > should assume that it sometimes gives "204, "Hardware Failure --
>> Flatness
>> > DAC Convergence: 198"
>> >
>> > i haven't opened the unit, and wasn't really planning on it. if
>> > there's something that you'd like to know that will be clear from just
>> > looking, and if it might appreciably change the unit's value for you,
>> > let me know and i'll find a screwdriver and take some more pictures.
>> >
>> > i have two original (i believe) copies of the manual. no service
>> manual.
>> > firmware version is "REV 2,1".
>> >
>> > so how about this: best offer (by private mail) under $350 gets it,
>> > shipping not included (it looks like shipping will be in the $40 to
>> $80
>> > range, once i find a box). for obvious reasons, a local-to-boston
>> > transaction will be given a lot of preference.
>> >
>> > i'll wait until next wednesday evening (i.e., about a week -- feb 13)
>> > before choosing a lucky recipient, since i hear some people don't read
>> > their mail as continuously as i do.
>> >
>> > paul
>> >
>> > i wrote:
>> > > thanks joe, and to the other couple of people who responded
>> privately.
>>
>> > > i need to figure out how much of a project i want this to be, and
>> how
>> > > it ranks against all the other projects that i also don't have time
>> > > for. :-)
>> > >
>> > > i have the full users manual (two copies!), but i don't have
>> anything
>> > > else that speaks GPIB. i'll play with some of the front panel
>> > > commands you've suggested, while i think about what to do. thanks!
>> > >
>> > > paul
>> > >
>> > > j. l. trantham wrote:
>> > > > Paul,
>> > > >
>> > > > First, the meter is worth something if nothing other than parts.
>> It is
>> > ...
>> > > > Good luck.
>> > > >
>> > > > Joe
>> > > >
>> > > > -----Original Message-----
>> > > > From: volt-nuts-bounces@febo.com
>> [mailto:volt-nuts-bounces@febo.com] On
>> > > > Behalf Of Paul Fox
>> > > > Sent: Monday, October 15, 2012 9:50 AM
>> > > > To: Discussion of precise voltage measurement
>> > > > Subject: [volt-nuts] hp 3458a advice
>> > > >
>> > > > hi --
>> > > >
>> > > > i'm not a volt-nuts regular, and not even a precision
>> measurement
>> nut
>> > > > at all, but given past discussion in the list archives, i think
>> you
>> > > > folks might be able to help.
>> > > >
>> > > > i was recently given an HP 3458a multimeter. the donor thought
>> it
>> > > > worked when he last used it, but that was quite a long time ago,
>> and i
>> > > > think it was stored in an attic for some time, so i'm not too
>> > > > surprised that it now doesn't pass self tests. :-/ he mainly
>> wanted
>> > > > it out of his house, along with a bunch of other vintage (sadly
>> mostly
>> > > > non-working) equipment he'd accumulated over the years.
>> > > >
>> > > > the self-test error is:
>> > > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
>> 198"
>> > > >
>> > > > judging from past list traffic, repairs on these units are
>> > > > expensive, and it would still need recalibration to be useful,
>> and
>> > > > there are may be other parts (eeproms?) that may not be good
>> > > > any more. i'm not really interested in pursuing any of this
>> > > > myself.
>> > > >
>> > > > what's worse is that the unit doesn't have a serial number. (i
>> assume
>> > > > the s/n would normally be obvious on the rear of the unit.) it
>> bears a
>> > > > sticker on the back that says "MTG PROTO 27". i think the unit
>> came
>> > > > from a place that themselves produced reference standards, and
>> they
>> > > > may have gotten a very early model from HP to work with. that's
>> > > > conjecture. i haven't opened the unit, so i don't know if there
>> are
>> > > > other clues to its age inside.
>> > > >
>> > > > so: is this machine useful to anyone? given the error message
>> above,
>> > > > are there any measurements at all that it might still get
>> correct,
>> in
>> > > > it's current state?
>> > > >
>> > > > paul
>> > > > =---------------------
>> > > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's
>> 63.1
>> degrees)
>> > > >
>> >
>> > =---------------------
>> > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0
>> degrees)
>> > _______________________________________________
>> > volt-nuts mailing list -- volt-nuts@febo.com
>> > To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > and follow the instructions there.
>>
>> =---------------------
>> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 19.2
>> degrees)
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
>
>
> --
> John Phillips
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
JP
John Phillips
Fri, Feb 8, 2013 3:48 AM
That is what I get for jumping in without reading the thread.
I have sent 2 or 3 units in for them to repair...
When they quit selling part I was forced to junk or have them repair.
What parts are needed?
On Thu, Feb 7, 2013 at 7:12 PM, Tom Miller tmiller11147@verizon.net wrote:
They might not take it in without a serial number.
Regards
----- Original Message ----- From: "John Phillips" <
john.phillips0@gmail.com>
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Thursday, February 07, 2013 9:34 PM
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
You should be able to any unit repaired by Agilent as long as all the
parts
are there for the $3000 flat rate repair and calibration charge.
On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham jltran@att.net wrote:
Paul,
The 'value' of a 3458A is based on several issues, not the least of which
is
supply and demand. You only have to look at theBay to get an idea of
what
units are selling for, both working and non-working. The fact that the
unit
passes it's 'Self Test' is not a guarantee that it would be able to be
calibrated to published performance specs.
Tha Agilent website is very useful in determining an 'upper bound' on the
value. The 3458A is currently orderable and currently supported. A new
3458A (without options) is available at $9,085 and would come with a
fresh
calibration and warranty.
Repair Per Incident is available for $2,660.64. This means that if you
have
a complete unit, they will repair it for that price. That also comes
with
a
fresh calibration. I don't know if that price includes all recommended
upgrades as listed in the Service Notes, of which there are 20, 6 of
which
list 'Modification Recommended' and one of which lists 'Modification
Available'. Perhaps someone on the forum knows the answer to this.
Certainly Gary Biermann at Agilent can answer this.
The fact that your unit does not have a serial number sticker but instead
'MTG PROTO 27' would be concerning as to whether Agilent would consider
this
a 'complete 3458A', and thus a candidate for repair, or, rather, a
'prototype' or some other descriptor that would imply that it looks like
a
3458A but is really not a 3458A.
Both new and newly calibrated units are candidates to add a Repair
Agreement, available for $178.68 per year with up to 5 year blocks
available
for purchase.
I note you mentioned firmware REV 2,1. The current firmware is 9,1.
Therefore it would appear that you have a rather old and probably not
'updated' unit. That plus the 'stored in an attic' history would be
concerning about the ability to repair the unit without changing several
assemblies, etc.
Assuming you could get the unit repaired and updated by Agilent, complete
with a fresh calibration, then the maximum value would be $9,085 minus
$2,660.64 minus what ever value you place on 'new' over 'old'.
Then you are left with an older unit, made in the USA, with no (or
perhaps
a
reduced) warranty as opposed to a new unit, made in Malaysia, with a
warranty. The question of whether a 'new' Malaysian unit or an 'old'
American unit is better is the subject of further discussion and
speculation.
If the unit cannot be repaired by Agilent, then, in my opinion, the value
goes down and approaches the value of the hardware, transformer, etc.,
with
serious questions about the state and reparability of the assemblies,
particularly given the 'stored in an attic' history.
Gary Biermann would be very helpful in getting the history and
reparability
of the unit. He would be worth contacting about these issues. He works
at
the Loveland Calibration Facility.
Hope this helps.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@**febo.comvolt-nuts-bounces@febo.com]
On
Behalf Of Paul Fox
Sent: Thursday, February 07, 2013 8:54 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to the
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure --
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from just
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't read
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded
i need to figure out how much of a project i want this to be, and
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than parts.
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision
at all, but given past discussion in the list archives, i think
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor thought
worked when he last used it, but that was quite a long time ago,
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly
it out of his house, along with a bunch of other vintage (sadly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful,
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i
the s/n would normally be obvious on the rear of the unit.) it
sticker on the back that says "MTG PROTO 27". i think the unit
from a place that themselves produced reference standards, and
may have gotten a very early model from HP to work with. that's
conjecture. i haven't opened the unit, so i don't know if there
other clues to its age inside.
so: is this machine useful to anyone? given the error message
are there any measurements at all that it might still get
and follow the instructions there.
That is what I get for jumping in without reading the thread.
I have sent 2 or 3 units in for them to repair...
When they quit selling part I was forced to junk or have them repair.
What parts are needed?
On Thu, Feb 7, 2013 at 7:12 PM, Tom Miller <tmiller11147@verizon.net> wrote:
> They might not take it in without a serial number.
>
> Regards
>
> ----- Original Message ----- From: "John Phillips" <
> john.phillips0@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Thursday, February 07, 2013 9:34 PM
>
> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>
>
> You should be able to any unit repaired by Agilent as long as all the
>> parts
>> are there for the $3000 flat rate repair and calibration charge.
>>
>> On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham <jltran@att.net> wrote:
>>
>> Paul,
>>>
>>> The 'value' of a 3458A is based on several issues, not the least of which
>>> is
>>> supply and demand. You only have to look at theBay to get an idea of
>>> what
>>> units are selling for, both working and non-working. The fact that the
>>> unit
>>> passes it's 'Self Test' is not a guarantee that it would be able to be
>>> calibrated to published performance specs.
>>>
>>> Tha Agilent website is very useful in determining an 'upper bound' on the
>>> value. The 3458A is currently orderable and currently supported. A new
>>> 3458A (without options) is available at $9,085 and would come with a
>>> fresh
>>> calibration and warranty.
>>>
>>> Repair Per Incident is available for $2,660.64. This means that if you
>>> have
>>> a complete unit, they will repair it for that price. That also comes
>>> with
>>> a
>>> fresh calibration. I don't know if that price includes all recommended
>>> upgrades as listed in the Service Notes, of which there are 20, 6 of
>>> which
>>> list 'Modification Recommended' and one of which lists 'Modification
>>> Available'. Perhaps someone on the forum knows the answer to this.
>>> Certainly Gary Biermann at Agilent can answer this.
>>>
>>> The fact that your unit does not have a serial number sticker but instead
>>> 'MTG PROTO 27' would be concerning as to whether Agilent would consider
>>> this
>>> a 'complete 3458A', and thus a candidate for repair, or, rather, a
>>> 'prototype' or some other descriptor that would imply that it looks like
>>> a
>>> 3458A but is really not a 3458A.
>>>
>>> Both new and newly calibrated units are candidates to add a Repair
>>> Agreement, available for $178.68 per year with up to 5 year blocks
>>> available
>>> for purchase.
>>>
>>> I note you mentioned firmware REV 2,1. The current firmware is 9,1.
>>> Therefore it would appear that you have a rather old and probably not
>>> 'updated' unit. That plus the 'stored in an attic' history would be
>>> concerning about the ability to repair the unit without changing several
>>> assemblies, etc.
>>>
>>> Assuming you could get the unit repaired and updated by Agilent, complete
>>> with a fresh calibration, then the maximum value would be $9,085 minus
>>> $2,660.64 minus what ever value you place on 'new' over 'old'.
>>>
>>> Then you are left with an older unit, made in the USA, with no (or
>>> perhaps
>>> a
>>> reduced) warranty as opposed to a new unit, made in Malaysia, with a
>>> warranty. The question of whether a 'new' Malaysian unit or an 'old'
>>> American unit is better is the subject of further discussion and
>>> speculation.
>>>
>>> If the unit cannot be repaired by Agilent, then, in my opinion, the value
>>> goes down and approaches the value of the hardware, transformer, etc.,
>>> with
>>> serious questions about the state and reparability of the assemblies,
>>> particularly given the 'stored in an attic' history.
>>>
>>> Gary Biermann would be very helpful in getting the history and
>>> reparability
>>> of the unit. He would be worth contacting about these issues. He works
>>> at
>>> the Loveland Calibration Facility.
>>>
>>> Hope this helps.
>>>
>>> Joe
>>>
>>> -----Original Message-----
>>> From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@**febo.com<volt-nuts-bounces@febo.com>]
>>> On
>>> Behalf Of Paul Fox
>>> Sent: Thursday, February 07, 2013 8:54 AM
>>> To: Discussion of precise voltage measurement
>>> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>>>
>>> part of me hates to do this, but... there have been some rule-breakers
>>> that ignored the "under $350" limit, and now they've ruined it for
>>> everyone. ;-)
>>>
>>> clearly the research i did into prices for "HP3458a, needs repair" was
>>> woefully inadequate.
>>>
>>> i guess the easiest thing to do is to simply remove the "under $350"
>>> part. that way everyone's working under the same assumptions. again,
>>> sorry. local buyers still get preference. (and, if you'd like to chew
>>> me out for being a bozo, that's fine, but please do it privately. maybe
>>> i should have used ebay in the first place. lesson learned.)
>>>
>>> paul
>>>
>>> i wrote:
>>> > resurrecting a somewhat older thread --
>>> >
>>> > it's now clear to me that i have little interest in (or time for)
>>> > resurrecting my 3458a. several people on the list were interested in
>>> > it when i first brought it up, and you were all very helpful. i'd
>>> > rather not go through the ebay hassle, and i'd rather make it
>>> > available to one of you folks first in any case.
>>> >
>>> > i've put up a small set of pictures, along with a short video of
>>> > the startup sequence at http://www.foxharp.net/hp3458a . re-reading
>>> > my original list message (below), i see that my startup video didn't
>>> > capture an error message i saw back in october. so in addition to the
>>> > "110, Calibration required -- ACAL" message in the video, i guess we
>>> > should assume that it sometimes gives "204, "Hardware Failure --
>>> Flatness
>>> > DAC Convergence: 198"
>>> >
>>> > i haven't opened the unit, and wasn't really planning on it. if
>>> > there's something that you'd like to know that will be clear from just
>>> > looking, and if it might appreciably change the unit's value for you,
>>> > let me know and i'll find a screwdriver and take some more pictures.
>>> >
>>> > i have two original (i believe) copies of the manual. no service
>>> manual.
>>> > firmware version is "REV 2,1".
>>> >
>>> > so how about this: best offer (by private mail) under $350 gets it,
>>> > shipping not included (it looks like shipping will be in the $40 to
>>> $80
>>> > range, once i find a box). for obvious reasons, a local-to-boston
>>> > transaction will be given a lot of preference.
>>> >
>>> > i'll wait until next wednesday evening (i.e., about a week -- feb 13)
>>> > before choosing a lucky recipient, since i hear some people don't read
>>> > their mail as continuously as i do.
>>> >
>>> > paul
>>> >
>>> > i wrote:
>>> > > thanks joe, and to the other couple of people who responded
>>> privately.
>>>
>>> > > i need to figure out how much of a project i want this to be, and
>>> how
>>> > > it ranks against all the other projects that i also don't have time
>>> > > for. :-)
>>> > >
>>> > > i have the full users manual (two copies!), but i don't have
>>> anything
>>> > > else that speaks GPIB. i'll play with some of the front panel
>>> > > commands you've suggested, while i think about what to do. thanks!
>>> > >
>>> > > paul
>>> > >
>>> > > j. l. trantham wrote:
>>> > > > Paul,
>>> > > >
>>> > > > First, the meter is worth something if nothing other than parts.
>>> It is
>>> > ...
>>> > > > Good luck.
>>> > > >
>>> > > > Joe
>>> > > >
>>> > > > -----Original Message-----
>>> > > > From: volt-nuts-bounces@febo.com
>>> [mailto:volt-nuts-bounces@**febo.com <volt-nuts-bounces@febo.com>] On
>>> > > > Behalf Of Paul Fox
>>> > > > Sent: Monday, October 15, 2012 9:50 AM
>>> > > > To: Discussion of precise voltage measurement
>>> > > > Subject: [volt-nuts] hp 3458a advice
>>> > > >
>>> > > > hi --
>>> > > >
>>> > > > i'm not a volt-nuts regular, and not even a precision
>>> measurement
>>> nut
>>> > > > at all, but given past discussion in the list archives, i think
>>> you
>>> > > > folks might be able to help.
>>> > > >
>>> > > > i was recently given an HP 3458a multimeter. the donor thought
>>> it
>>> > > > worked when he last used it, but that was quite a long time ago,
>>> and i
>>> > > > think it was stored in an attic for some time, so i'm not too
>>> > > > surprised that it now doesn't pass self tests. :-/ he mainly
>>> wanted
>>> > > > it out of his house, along with a bunch of other vintage (sadly
>>> mostly
>>> > > > non-working) equipment he'd accumulated over the years.
>>> > > >
>>> > > > the self-test error is:
>>> > > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
>>> 198"
>>> > > >
>>> > > > judging from past list traffic, repairs on these units are
>>> > > > expensive, and it would still need recalibration to be useful,
>>> and
>>> > > > there are may be other parts (eeproms?) that may not be good
>>> > > > any more. i'm not really interested in pursuing any of this
>>> > > > myself.
>>> > > >
>>> > > > what's worse is that the unit doesn't have a serial number. (i
>>> assume
>>> > > > the s/n would normally be obvious on the rear of the unit.) it
>>> bears a
>>> > > > sticker on the back that says "MTG PROTO 27". i think the unit
>>> came
>>> > > > from a place that themselves produced reference standards, and
>>> they
>>> > > > may have gotten a very early model from HP to work with. that's
>>> > > > conjecture. i haven't opened the unit, so i don't know if there
>>> are
>>> > > > other clues to its age inside.
>>> > > >
>>> > > > so: is this machine useful to anyone? given the error message
>>> above,
>>> > > > are there any measurements at all that it might still get
>>> correct,
>>> in
>>> > > > it's current state?
>>> > > >
>>> > > > paul
>>> > > > =---------------------
>>> > > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's
>>> 63.1
>>> degrees)
>>> > > >
>>> >
>>> > =---------------------
>>> > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0
>>> degrees)
>>> > ______________________________**_________________
>>> > volt-nuts mailing list -- volt-nuts@febo.com
>>> > To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> > and follow the instructions there.
>>>
>>> =---------------------
>>> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 19.2
>>> degrees)
>>> ______________________________**_________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> and follow the instructions there.
>>>
>>> ______________________________**_________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> and follow the instructions there.
>>>
>>>
>>
>>
>> --
>> John Phillips
>> ______________________________**_________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>> and follow the instructions there.
>>
>>
> ______________________________**_________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/**
> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
> and follow the instructions there.
>
--
John Phillips
JP
John Phillips
Fri, Feb 8, 2013 4:10 AM
I read most of the thread. This could be a pre-production unit. The errors
indicate it could need both the in guard and out guard CPUs. The in guard
unit (03458-05503) I think is involved with the A/D conversion and that is
where the flatness error is most likly coming from.
On Thu, Feb 7, 2013 at 7:48 PM, John Phillips john.phillips0@gmail.comwrote:
That is what I get for jumping in without reading the thread.
I have sent 2 or 3 units in for them to repair...
When they quit selling part I was forced to junk or have them repair.
What parts are needed?
on Thu, Feb 7, 2013 at 7:12 PM, Tom Miller tmiller11147@verizon.netwrte:
They might not take it in without a serial number.
Regards
----- Original Message ----- From: "John Phillips" <
john.phillips0@gmail.com>
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Thursday, February 07, 2013 9:34 PM
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
You should be able to any unit repaired by Agilent as long as all the
parts
are there for the $3000 flat rate repair and calibration charge.
On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham jltran@att.net wrote:
Paul,
The 'value' of a 3458A is based on several issues, not the least of
which
is
supply and demand. You only have to look at theBay to get an idea of
what
units are selling for, both working and non-working. The fact that the
unit
passes it's 'Self Test' is not a guarantee that it would be able to be
calibrated to published performance specs.
Tha Agilent website is very useful in determining an 'upper bound' on
the
value. The 3458A is currently orderable and currently supported. A new
3458A (without options) is available at $9,085 and would come with a
fresh
calibration and warranty.
Repair Per Incident is available for $2,660.64. This means that if you
have
a complete unit, they will repair it for that price. That also comes
with
a
fresh calibration. I don't know if that price includes all recommended
upgrades as listed in the Service Notes, of which there are 20, 6 of
which
list 'Modification Recommended' and one of which lists 'Modification
Available'. Perhaps someone on the forum knows the answer to this.
Certainly Gary Biermann at Agilent can answer this.
The fact that your unit does not have a serial number sticker but
instead
'MTG PROTO 27' would be concerning as to whether Agilent would consider
this
a 'complete 3458A', and thus a candidate for repair, or, rather, a
'prototype' or some other descriptor that would imply that it looks
like a
3458A but is really not a 3458A.
Both new and newly calibrated units are candidates to add a Repair
Agreement, available for $178.68 per year with up to 5 year blocks
available
for purchase.
I note you mentioned firmware REV 2,1. The current firmware is 9,1.
Therefore it would appear that you have a rather old and probably not
'updated' unit. That plus the 'stored in an attic' history would be
concerning about the ability to repair the unit without changing several
assemblies, etc.
Assuming you could get the unit repaired and updated by Agilent,
complete
with a fresh calibration, then the maximum value would be $9,085 minus
$2,660.64 minus what ever value you place on 'new' over 'old'.
Then you are left with an older unit, made in the USA, with no (or
perhaps
a
reduced) warranty as opposed to a new unit, made in Malaysia, with a
warranty. The question of whether a 'new' Malaysian unit or an 'old'
American unit is better is the subject of further discussion and
speculation.
If the unit cannot be repaired by Agilent, then, in my opinion, the
value
goes down and approaches the value of the hardware, transformer, etc.,
with
serious questions about the state and reparability of the assemblies,
particularly given the 'stored in an attic' history.
Gary Biermann would be very helpful in getting the history and
reparability
of the unit. He would be worth contacting about these issues. He
works at
the Loveland Calibration Facility.
Hope this helps.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@**febo.comvolt-nuts-bounces@febo.com]
On
Behalf Of Paul Fox
Sent: Thursday, February 07, 2013 8:54 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
part of me hates to do this, but... there have been some rule-breakers
that ignored the "under $350" limit, and now they've ruined it for
everyone. ;-)
clearly the research i did into prices for "HP3458a, needs repair" was
woefully inadequate.
i guess the easiest thing to do is to simply remove the "under $350"
part. that way everyone's working under the same assumptions. again,
sorry. local buyers still get preference. (and, if you'd like to chew
me out for being a bozo, that's fine, but please do it privately. maybe
i should have used ebay in the first place. lesson learned.)
paul
i wrote:
resurrecting a somewhat older thread --
it's now clear to me that i have little interest in (or time for)
resurrecting my 3458a. several people on the list were interested in
it when i first brought it up, and you were all very helpful. i'd
rather not go through the ebay hassle, and i'd rather make it
available to one of you folks first in any case.
i've put up a small set of pictures, along with a short video of
the startup sequence at http://www.foxharp.net/hp3458a . re-reading
my original list message (below), i see that my startup video didn't
capture an error message i saw back in october. so in addition to
"110, Calibration required -- ACAL" message in the video, i guess we
should assume that it sometimes gives "204, "Hardware Failure --
DAC Convergence: 198"
i haven't opened the unit, and wasn't really planning on it. if
there's something that you'd like to know that will be clear from
looking, and if it might appreciably change the unit's value for you,
let me know and i'll find a screwdriver and take some more pictures.
i have two original (i believe) copies of the manual. no service
firmware version is "REV 2,1".
so how about this: best offer (by private mail) under $350 gets it,
shipping not included (it looks like shipping will be in the $40 to
range, once i find a box). for obvious reasons, a local-to-boston
transaction will be given a lot of preference.
i'll wait until next wednesday evening (i.e., about a week -- feb 13)
before choosing a lucky recipient, since i hear some people don't
their mail as continuously as i do.
paul
i wrote:
thanks joe, and to the other couple of people who responded
i need to figure out how much of a project i want this to be, and
it ranks against all the other projects that i also don't have
for. :-)
i have the full users manual (two copies!), but i don't have
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do.
paul
j. l. trantham wrote:
Paul,
First, the meter is worth something if nothing other than
Behalf Of Paul Fox
Sent: Monday, October 15, 2012 9:50 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] hp 3458a advice
hi --
i'm not a volt-nuts regular, and not even a precision
at all, but given past discussion in the list archives, i think
folks might be able to help.
i was recently given an HP 3458a multimeter. the donor
worked when he last used it, but that was quite a long time
think it was stored in an attic for some time, so i'm not too
surprised that it now doesn't pass self tests. :-/ he mainly
it out of his house, along with a bunch of other vintage (sadly
non-working) equipment he'd accumulated over the years.
the self-test error is:
ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
judging from past list traffic, repairs on these units are
expensive, and it would still need recalibration to be useful,
there are may be other parts (eeproms?) that may not be good
any more. i'm not really interested in pursuing any of this
myself.
what's worse is that the unit doesn't have a serial number. (i
the s/n would normally be obvious on the rear of the unit.) it
sticker on the back that says "MTG PROTO 27". i think the unit
from a place that themselves produced reference standards, and
may have gotten a very early model from HP to work with.
conjecture. i haven't opened the unit, so i don't know if
other clues to its age inside.
so: is this machine useful to anyone? given the error message
are there any measurements at all that it might still get
and follow the instructions there.
I read most of the thread. This could be a pre-production unit. The errors
indicate it could need both the in guard and out guard CPUs. The in guard
unit (03458-05503) I think is involved with the A/D conversion and that is
where the flatness error is most likly coming from.
On Thu, Feb 7, 2013 at 7:48 PM, John Phillips <john.phillips0@gmail.com>wrote:
> That is what I get for jumping in without reading the thread.
> I have sent 2 or 3 units in for them to repair...
> When they quit selling part I was forced to junk or have them repair.
> What parts are needed?
>
>
> on Thu, Feb 7, 2013 at 7:12 PM, Tom Miller <tmiller11147@verizon.net>wrte:
>
>> They might not take it in without a serial number.
>>
>> Regards
>>
>> ----- Original Message ----- From: "John Phillips" <
>> john.phillips0@gmail.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Sent: Thursday, February 07, 2013 9:34 PM
>>
>> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>>
>>
>> You should be able to any unit repaired by Agilent as long as all the
>>> parts
>>> are there for the $3000 flat rate repair and calibration charge.
>>>
>>> On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham <jltran@att.net> wrote:
>>>
>>> Paul,
>>>>
>>>> The 'value' of a 3458A is based on several issues, not the least of
>>>> which
>>>> is
>>>> supply and demand. You only have to look at theBay to get an idea of
>>>> what
>>>> units are selling for, both working and non-working. The fact that the
>>>> unit
>>>> passes it's 'Self Test' is not a guarantee that it would be able to be
>>>> calibrated to published performance specs.
>>>>
>>>> Tha Agilent website is very useful in determining an 'upper bound' on
>>>> the
>>>> value. The 3458A is currently orderable and currently supported. A new
>>>> 3458A (without options) is available at $9,085 and would come with a
>>>> fresh
>>>> calibration and warranty.
>>>>
>>>> Repair Per Incident is available for $2,660.64. This means that if you
>>>> have
>>>> a complete unit, they will repair it for that price. That also comes
>>>> with
>>>> a
>>>> fresh calibration. I don't know if that price includes all recommended
>>>> upgrades as listed in the Service Notes, of which there are 20, 6 of
>>>> which
>>>> list 'Modification Recommended' and one of which lists 'Modification
>>>> Available'. Perhaps someone on the forum knows the answer to this.
>>>> Certainly Gary Biermann at Agilent can answer this.
>>>>
>>>> The fact that your unit does not have a serial number sticker but
>>>> instead
>>>> 'MTG PROTO 27' would be concerning as to whether Agilent would consider
>>>> this
>>>> a 'complete 3458A', and thus a candidate for repair, or, rather, a
>>>> 'prototype' or some other descriptor that would imply that it looks
>>>> like a
>>>> 3458A but is really not a 3458A.
>>>>
>>>> Both new and newly calibrated units are candidates to add a Repair
>>>> Agreement, available for $178.68 per year with up to 5 year blocks
>>>> available
>>>> for purchase.
>>>>
>>>> I note you mentioned firmware REV 2,1. The current firmware is 9,1.
>>>> Therefore it would appear that you have a rather old and probably not
>>>> 'updated' unit. That plus the 'stored in an attic' history would be
>>>> concerning about the ability to repair the unit without changing several
>>>> assemblies, etc.
>>>>
>>>> Assuming you could get the unit repaired and updated by Agilent,
>>>> complete
>>>> with a fresh calibration, then the maximum value would be $9,085 minus
>>>> $2,660.64 minus what ever value you place on 'new' over 'old'.
>>>>
>>>> Then you are left with an older unit, made in the USA, with no (or
>>>> perhaps
>>>> a
>>>> reduced) warranty as opposed to a new unit, made in Malaysia, with a
>>>> warranty. The question of whether a 'new' Malaysian unit or an 'old'
>>>> American unit is better is the subject of further discussion and
>>>> speculation.
>>>>
>>>> If the unit cannot be repaired by Agilent, then, in my opinion, the
>>>> value
>>>> goes down and approaches the value of the hardware, transformer, etc.,
>>>> with
>>>> serious questions about the state and reparability of the assemblies,
>>>> particularly given the 'stored in an attic' history.
>>>>
>>>> Gary Biermann would be very helpful in getting the history and
>>>> reparability
>>>> of the unit. He would be worth contacting about these issues. He
>>>> works at
>>>> the Loveland Calibration Facility.
>>>>
>>>> Hope this helps.
>>>>
>>>> Joe
>>>>
>>>> -----Original Message-----
>>>> From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@**febo.com<volt-nuts-bounces@febo.com>]
>>>> On
>>>> Behalf Of Paul Fox
>>>> Sent: Thursday, February 07, 2013 8:54 AM
>>>> To: Discussion of precise voltage measurement
>>>> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>>>>
>>>> part of me hates to do this, but... there have been some rule-breakers
>>>> that ignored the "under $350" limit, and now they've ruined it for
>>>> everyone. ;-)
>>>>
>>>> clearly the research i did into prices for "HP3458a, needs repair" was
>>>> woefully inadequate.
>>>>
>>>> i guess the easiest thing to do is to simply remove the "under $350"
>>>> part. that way everyone's working under the same assumptions. again,
>>>> sorry. local buyers still get preference. (and, if you'd like to chew
>>>> me out for being a bozo, that's fine, but please do it privately. maybe
>>>> i should have used ebay in the first place. lesson learned.)
>>>>
>>>> paul
>>>>
>>>> i wrote:
>>>> > resurrecting a somewhat older thread --
>>>> >
>>>> > it's now clear to me that i have little interest in (or time for)
>>>> > resurrecting my 3458a. several people on the list were interested in
>>>> > it when i first brought it up, and you were all very helpful. i'd
>>>> > rather not go through the ebay hassle, and i'd rather make it
>>>> > available to one of you folks first in any case.
>>>> >
>>>> > i've put up a small set of pictures, along with a short video of
>>>> > the startup sequence at http://www.foxharp.net/hp3458a . re-reading
>>>> > my original list message (below), i see that my startup video didn't
>>>> > capture an error message i saw back in october. so in addition to
>>>> the
>>>> > "110, Calibration required -- ACAL" message in the video, i guess we
>>>> > should assume that it sometimes gives "204, "Hardware Failure --
>>>> Flatness
>>>> > DAC Convergence: 198"
>>>> >
>>>> > i haven't opened the unit, and wasn't really planning on it. if
>>>> > there's something that you'd like to know that will be clear from
>>>> just
>>>> > looking, and if it might appreciably change the unit's value for you,
>>>> > let me know and i'll find a screwdriver and take some more pictures.
>>>> >
>>>> > i have two original (i believe) copies of the manual. no service
>>>> manual.
>>>> > firmware version is "REV 2,1".
>>>> >
>>>> > so how about this: best offer (by private mail) under $350 gets it,
>>>> > shipping not included (it looks like shipping will be in the $40 to
>>>> $80
>>>> > range, once i find a box). for obvious reasons, a local-to-boston
>>>> > transaction will be given a lot of preference.
>>>> >
>>>> > i'll wait until next wednesday evening (i.e., about a week -- feb 13)
>>>> > before choosing a lucky recipient, since i hear some people don't
>>>> read
>>>> > their mail as continuously as i do.
>>>> >
>>>> > paul
>>>> >
>>>> > i wrote:
>>>> > > thanks joe, and to the other couple of people who responded
>>>> privately.
>>>>
>>>> > > i need to figure out how much of a project i want this to be, and
>>>> how
>>>> > > it ranks against all the other projects that i also don't have
>>>> time
>>>> > > for. :-)
>>>> > >
>>>> > > i have the full users manual (two copies!), but i don't have
>>>> anything
>>>> > > else that speaks GPIB. i'll play with some of the front panel
>>>> > > commands you've suggested, while i think about what to do.
>>>> thanks!
>>>> > >
>>>> > > paul
>>>> > >
>>>> > > j. l. trantham wrote:
>>>> > > > Paul,
>>>> > > >
>>>> > > > First, the meter is worth something if nothing other than
>>>> parts.
>>>> It is
>>>> > ...
>>>> > > > Good luck.
>>>> > > >
>>>> > > > Joe
>>>> > > >
>>>> > > > -----Original Message-----
>>>> > > > From: volt-nuts-bounces@febo.com
>>>> [mailto:volt-nuts-bounces@**febo.com <volt-nuts-bounces@febo.com>] On
>>>> > > > Behalf Of Paul Fox
>>>> > > > Sent: Monday, October 15, 2012 9:50 AM
>>>> > > > To: Discussion of precise voltage measurement
>>>> > > > Subject: [volt-nuts] hp 3458a advice
>>>> > > >
>>>> > > > hi --
>>>> > > >
>>>> > > > i'm not a volt-nuts regular, and not even a precision
>>>> measurement
>>>> nut
>>>> > > > at all, but given past discussion in the list archives, i think
>>>> you
>>>> > > > folks might be able to help.
>>>> > > >
>>>> > > > i was recently given an HP 3458a multimeter. the donor
>>>> thought it
>>>> > > > worked when he last used it, but that was quite a long time
>>>> ago,
>>>> and i
>>>> > > > think it was stored in an attic for some time, so i'm not too
>>>> > > > surprised that it now doesn't pass self tests. :-/ he mainly
>>>> wanted
>>>> > > > it out of his house, along with a bunch of other vintage (sadly
>>>> mostly
>>>> > > > non-working) equipment he'd accumulated over the years.
>>>> > > >
>>>> > > > the self-test error is:
>>>> > > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
>>>> 198"
>>>> > > >
>>>> > > > judging from past list traffic, repairs on these units are
>>>> > > > expensive, and it would still need recalibration to be useful,
>>>> and
>>>> > > > there are may be other parts (eeproms?) that may not be good
>>>> > > > any more. i'm not really interested in pursuing any of this
>>>> > > > myself.
>>>> > > >
>>>> > > > what's worse is that the unit doesn't have a serial number. (i
>>>> assume
>>>> > > > the s/n would normally be obvious on the rear of the unit.) it
>>>> bears a
>>>> > > > sticker on the back that says "MTG PROTO 27". i think the unit
>>>> came
>>>> > > > from a place that themselves produced reference standards, and
>>>> they
>>>> > > > may have gotten a very early model from HP to work with.
>>>> that's
>>>> > > > conjecture. i haven't opened the unit, so i don't know if
>>>> there
>>>> are
>>>> > > > other clues to its age inside.
>>>> > > >
>>>> > > > so: is this machine useful to anyone? given the error message
>>>> above,
>>>> > > > are there any measurements at all that it might still get
>>>> correct,
>>>> in
>>>> > > > it's current state?
>>>> > > >
>>>> > > > paul
>>>> > > > =---------------------
>>>> > > > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's
>>>> 63.1
>>>> degrees)
>>>> > > >
>>>> >
>>>> > =---------------------
>>>> > paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 30.0
>>>> degrees)
>>>> > ______________________________**_________________
>>>> > volt-nuts mailing list -- volt-nuts@febo.com
>>>> > To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> > and follow the instructions there.
>>>>
>>>> =---------------------
>>>> paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 19.2
>>>> degrees)
>>>> ______________________________**_________________
>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> and follow the instructions there.
>>>>
>>>> ______________________________**_________________
>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> and follow the instructions there.
>>>>
>>>>
>>>
>>>
>>> --
>>> John Phillips
>>> ______________________________**_________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> and follow the instructions there.
>>>
>>>
>> ______________________________**_________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>> and follow the instructions there.
>>
>
>
>
> --
> John Phillips
--
John Phillips
PF
Paul Fox
Thu, Feb 14, 2013 9:29 PM
hi -- just to close the loop -- i've found a good home for my
meter locally. if this falls through, i'll contact "the next in
line", so to speak, but i think i'm set.
thanks for all the interest and information -- i really do appreciate it!
paul
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 36.5 degrees)
hi -- just to close the loop -- i've found a good home for my
meter locally. if this falls through, i'll contact "the next in
line", so to speak, but i think i'm set.
thanks for all the interest and information -- i really do appreciate it!
paul
=---------------------
paul fox, pgf@foxharp.boston.ma.us (arlington, ma, where it's 36.5 degrees)