I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is the best resolution they can provide for a one second pulse (over one period)?

Do you have any recommendation as to which one of the three models is the better unit?  (Their display technology, age and cost not withstanding.)  All three models seem greatly overstated in price on eBay, seems, too.  Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Don Lewis
Austin, TX (Hyde Park)

I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution. In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period. The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period). I was looking at the HP5370/ 71/ 72 series counter/ timers. What is the best resolution they can provide for a one second pulse (over one period)? Do you have any recommendation as to which one of the three models is the better unit? (Their display technology, age and cost not withstanding.) All three models seem greatly overstated in price on eBay, seems, too. Not sure why, ...just supply and demand; I presume. Thank you for your advice; I appreciate your thoughts. Don Lewis Austin, TX (Hyde Park)
MD
Magnus Danielson
Wed, Jan 9, 2013 1:42 AM

Hi Don,

On 01/09/2013 02:32 AM, dlewis6767 wrote:

I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is the best resolution they can provide for a one second pulse (over one period)?

Do you have any recommendation as to which one of the three models is the better unit?  (Their display technology, age and cost not withstanding.)  All three models seem greatly overstated in price on eBay, seems, too.  Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Of the given set, the HP5370 will have much better single shot
resolution compared to the HP5371/5372. The later is better at recording
high rate events.

I have both HP5370 and HP5372 as they fill different purposes. I also
have counters of better performance than these.

Trouble is, with PPS signals you experience jitter such that a 1 ns
resolution of a HP5335 will suffice for most usages.

Cheers,
Magnus

Hi Don, On 01/09/2013 02:32 AM, dlewis6767 wrote: > I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution. > > In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period. > > The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period). > > I was looking at the HP5370/ 71/ 72 series counter/ timers. What is the best resolution they can provide for a one second pulse (over one period)? > > Do you have any recommendation as to which one of the three models is the better unit? (Their display technology, age and cost not withstanding.) All three models seem greatly overstated in price on eBay, seems, too. Not sure why, ...just supply and demand; I presume. > > Thank you for your advice; I appreciate your thoughts. Of the given set, the HP5370 will have much better single shot resolution compared to the HP5371/5372. The later is better at recording high rate events. I have both HP5370 and HP5372 as they fill different purposes. I also have counters of better performance than these. Trouble is, with PPS signals you experience jitter such that a 1 ns resolution of a HP5335 will suffice for most usages. Cheers, Magnus
D
dlewis6767
Wed, Jan 9, 2013 2:50 AM

Hi Magnus, ....good day.

Thanks for the thoughts, ...that was pretty much where I was headed.  I
think I will hold out a while for a 5370, ....watch eBay, see if a
reasonably priced unit pops up.

If anyone has one for sale for a nice price, please let me know.

-Don


From: "Magnus Danielson" magnus@rubidium.dyndns.org
Sent: Tuesday, January 08, 2013 7:42 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Interval Timer Recommendation

Hi Don,

On 01/09/2013 02:32 AM, dlewis6767 wrote:

I am looking to upgrade my older HP 5245L counter/ timer to one with
greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers) and
just want to start measuring their 1PPS jitter, etc. over one period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period
at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is the
best resolution they can provide for a one second pulse (over one
period)?

Do you have any recommendation as to which one of the three models is the
better unit?  (Their display technology, age and cost not withstanding.)
All three models seem greatly overstated in price on eBay, seems, too.
Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Of the given set, the HP5370 will have much better single shot resolution
compared to the HP5371/5372. The later is better at recording high rate
events.

I have both HP5370 and HP5372 as they fill different purposes. I also have
counters of better performance than these.

Trouble is, with PPS signals you experience jitter such that a 1 ns
resolution of a HP5335 will suffice for most usages.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Magnus, ....good day. Thanks for the thoughts, ...that was pretty much where I was headed. I think I will hold out a while for a 5370, ....watch eBay, see if a reasonably priced unit pops up. If anyone has one for sale for a nice price, please let me know. -Don -------------------------------------------------- From: "Magnus Danielson" <magnus@rubidium.dyndns.org> Sent: Tuesday, January 08, 2013 7:42 PM To: <time-nuts@febo.com> Subject: Re: [time-nuts] Interval Timer Recommendation > Hi Don, > > On 01/09/2013 02:32 AM, dlewis6767 wrote: >> I am looking to upgrade my older HP 5245L counter/ timer to one with >> greater resolution. >> >> In my home lab, I have been collecting GPS timing modules (receivers) and >> just want to start measuring their 1PPS jitter, etc. over one period. >> >> The 5245L will only measure to 100ns with a 1 second pulse (.1us period >> at 1 period). >> >> I was looking at the HP5370/ 71/ 72 series counter/ timers. What is the >> best resolution they can provide for a one second pulse (over one >> period)? >> >> Do you have any recommendation as to which one of the three models is the >> better unit? (Their display technology, age and cost not withstanding.) >> All three models seem greatly overstated in price on eBay, seems, too. >> Not sure why, ...just supply and demand; I presume. >> >> Thank you for your advice; I appreciate your thoughts. > > Of the given set, the HP5370 will have much better single shot resolution > compared to the HP5371/5372. The later is better at recording high rate > events. > > I have both HP5370 and HP5372 as they fill different purposes. I also have > counters of better performance than these. > > Trouble is, with PPS signals you experience jitter such that a 1 ns > resolution of a HP5335 will suffice for most usages. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
EP
Ed Palmer
Wed, Jan 9, 2013 4:39 AM

In the past there has been much discussion on this list regarding the
relative strengths and weaknesses of the 5370 (20ps resolution) and the
5371/5372 (150ps resolution) units.  There are other units that have
even better resolution, but they tend to be pricey and/or more specialized.

Moving down another step in resolution, the Racal-Dana 1991 and 1992 (1
ns resolution) counters also have their fans (like me), but they do have
problems with their front panel buttons (actually the switches under the
buttons).

Yes, the prices on ebay are becoming increasingly ridiculous.
Occasionally a seller will have an attack of sanity and post a
reasonable price.  You may also get lucky with a unit that is defective,
or advertised as such, but has either no fault or a repairable one.

More info on all of these topics is available by using Google with the
site:www.febo.com advanced search term.

Ed

On 1/8/2013 7:32 PM, dlewis6767 wrote:

I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is the best resolution they can provide for a one second pulse (over one period)?

Do you have any recommendation as to which one of the three models is the better unit?  (Their display technology, age and cost not withstanding.)  All three models seem greatly overstated in price on eBay, seems, too.  Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Don Lewis
Austin, TX (Hyde Park)

In the past there has been much discussion on this list regarding the relative strengths and weaknesses of the 5370 (20ps resolution) and the 5371/5372 (150ps resolution) units. There are other units that have even better resolution, but they tend to be pricey and/or more specialized. Moving down another step in resolution, the Racal-Dana 1991 and 1992 (1 ns resolution) counters also have their fans (like me), but they do have problems with their front panel buttons (actually the switches under the buttons). Yes, the prices on ebay are becoming increasingly ridiculous. Occasionally a seller will have an attack of sanity and post a reasonable price. You may also get lucky with a unit that is defective, or advertised as such, but has either no fault or a repairable one. More info on all of these topics is available by using Google with the site:www.febo.com advanced search term. Ed On 1/8/2013 7:32 PM, dlewis6767 wrote: > I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution. > > In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period. > > The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period). > > I was looking at the HP5370/ 71/ 72 series counter/ timers. What is the best resolution they can provide for a one second pulse (over one period)? > > Do you have any recommendation as to which one of the three models is the better unit? (Their display technology, age and cost not withstanding.) All three models seem greatly overstated in price on eBay, seems, too. Not sure why, ...just supply and demand; I presume. > > Thank you for your advice; I appreciate your thoughts. > > Don Lewis > Austin, TX (Hyde Park) >
CA
Chris Albertson
Wed, Jan 9, 2013 4:54 AM

Wasn't this the exact application the PicTic was developed for "So one
would not have to tie up an expensive counter for something like comparing
two standards."  It would in fact be perfect.  You'd need to build several
of them one for every two GPSes

In my home lab, I have been collecting GPS timing modules (receivers) and

just want to start measuring their 1PPS jitter, etc. over one period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period
at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is the
best resolution they can provide for a one second pulse (over one period)?

--

Chris Albertson
Redondo Beach, California

Wasn't this the exact application the PicTic was developed for "So one would not have to tie up an expensive counter for something like comparing two standards." It would in fact be perfect. You'd need to build several of them one for every two GPSes > In my home lab, I have been collecting GPS timing modules (receivers) and >>> just want to start measuring their 1PPS jitter, etc. over one period. >>> >>> The 5245L will only measure to 100ns with a 1 second pulse (.1us period >>> at 1 period). >>> >>> I was looking at the HP5370/ 71/ 72 series counter/ timers. What is the >>> best resolution they can provide for a one second pulse (over one period)? >> >> -- Chris Albertson Redondo Beach, California
BC
Bob Camp
Wed, Jan 9, 2013 12:43 PM

Hi

There are a number of counters that are a lot better than what you have. There's also a limited amount of money in the universe. I'd suggest setting a rational budget and time frame. This is all used gear. You might get a perfect one that never ever breaks. More likely you will need to get a couple and swap things around. I'd plan on buying 3 or 4 with what ever the budget is. If you get lucky, you come in under budget.

Less fancy:

5334 - couple of ns resolution, nice and small, B is very different than the A so get one or the other
5335 - faster bigger heavier, one ns resolution.

both should be around $100 before shipping.

Very cool:

53131 500 ps resolution
53132 150 ps resolution

both have displays that can wear out. They are no longer supported by HP so you might find somebody dumping them. Hard to find for less than $800.

Big and heavy:

5371, 5372 same resolution as the 53132

They don't show up much. I've seen them for $200. You probably should expect to pay $400. Expect shipping to be non-trivial.

High end:

SR620
5370
53230a

All are in the 20 ps range. You do see 5370's for $200. More commonly you see them for quite a bit more. The 620's show up anywhere between $800 and $3000. The 53230a is brand new with a warranty from Agilent for about $3700.

That's just one set of parts. There are others. There are more issues than just price to consider. The 5371 is a big piece of gear. The 53xxx boxes are all small enough to carry around.

Lots of choices.

Bob

On Jan 8, 2013, at 9:50 PM, dlewis6767 dlewis6767@austin.rr.com wrote:

Hi Magnus, ....good day.

Thanks for the thoughts, ...that was pretty much where I was headed.  I think I will hold out a while for a 5370, ....watch eBay, see if a reasonably priced unit pops up.

If anyone has one for sale for a nice price, please let me know.

-Don


From: "Magnus Danielson" magnus@rubidium.dyndns.org
Sent: Tuesday, January 08, 2013 7:42 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Interval Timer Recommendation

Hi Don,

On 01/09/2013 02:32 AM, dlewis6767 wrote:

I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is the best resolution they can provide for a one second pulse (over one period)?

Do you have any recommendation as to which one of the three models is the better unit?  (Their display technology, age and cost not withstanding.) All three models seem greatly overstated in price on eBay, seems, too. Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Of the given set, the HP5370 will have much better single shot resolution compared to the HP5371/5372. The later is better at recording high rate events.

I have both HP5370 and HP5372 as they fill different purposes. I also have counters of better performance than these.

Trouble is, with PPS signals you experience jitter such that a 1 ns resolution of a HP5335 will suffice for most usages.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi There are a number of counters that are a lot better than what you have. There's also a limited amount of money in the universe. I'd suggest setting a rational budget and time frame. This is all used gear. You *might* get a perfect one that never ever breaks. More likely you will need to get a couple and swap things around. I'd plan on buying 3 or 4 with what ever the budget is. If you get lucky, you come in under budget. Less fancy: 5334 - couple of ns resolution, nice and small, B is *very* different than the A so get one or the other 5335 - faster bigger heavier, one ns resolution. both should be around $100 before shipping. Very cool: 53131 500 ps resolution 53132 150 ps resolution both have displays that can wear out. They are no longer supported by HP so you *might* find somebody dumping them. Hard to find for less than $800. Big and heavy: 5371, 5372 same resolution as the 53132 They don't show up much. I've seen them for $200. You probably should expect to pay $400. Expect shipping to be non-trivial. High end: SR620 5370 53230a All are in the 20 ps range. You do see 5370's for $200. More commonly you see them for quite a bit more. The 620's show up anywhere between $800 and $3000. The 53230a is brand new with a warranty from Agilent for about $3700. That's just one set of parts. There are others. There are more issues than just price to consider. The 5371 is a *big* piece of gear. The 53xxx boxes are all small enough to carry around. Lots of choices. Bob On Jan 8, 2013, at 9:50 PM, dlewis6767 <dlewis6767@austin.rr.com> wrote: > Hi Magnus, ....good day. > > Thanks for the thoughts, ...that was pretty much where I was headed. I think I will hold out a while for a 5370, ....watch eBay, see if a reasonably priced unit pops up. > > If anyone has one for sale for a nice price, please let me know. > > -Don > > > > > > > -------------------------------------------------- > From: "Magnus Danielson" <magnus@rubidium.dyndns.org> > Sent: Tuesday, January 08, 2013 7:42 PM > To: <time-nuts@febo.com> > Subject: Re: [time-nuts] Interval Timer Recommendation > >> Hi Don, >> >> On 01/09/2013 02:32 AM, dlewis6767 wrote: >>> I am looking to upgrade my older HP 5245L counter/ timer to one with greater resolution. >>> >>> In my home lab, I have been collecting GPS timing modules (receivers) and just want to start measuring their 1PPS jitter, etc. over one period. >>> >>> The 5245L will only measure to 100ns with a 1 second pulse (.1us period at 1 period). >>> >>> I was looking at the HP5370/ 71/ 72 series counter/ timers. What is the best resolution they can provide for a one second pulse (over one period)? >>> >>> Do you have any recommendation as to which one of the three models is the better unit? (Their display technology, age and cost not withstanding.) All three models seem greatly overstated in price on eBay, seems, too. Not sure why, ...just supply and demand; I presume. >>> >>> Thank you for your advice; I appreciate your thoughts. >> >> Of the given set, the HP5370 will have much better single shot resolution compared to the HP5371/5372. The later is better at recording high rate events. >> >> I have both HP5370 and HP5372 as they fill different purposes. I also have counters of better performance than these. >> >> Trouble is, with PPS signals you experience jitter such that a 1 ns resolution of a HP5335 will suffice for most usages. >> >> Cheers, >> Magnus >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
D
dlewis6767
Wed, Jan 9, 2013 1:06 PM

Thank you Bob, Ed, Magnus , ..and all.

Excellent reviews, ...detailed and clear.  I REALLY do appreciate it.

HP has turned out so much equipment over the years, it is often good to see
a comparison.  This was excellent.

I'll wait for a good deal, ....and don the road bid alongside many of you ,
I presume.  :-)

I may opt for one of the better units, ...as this disease progresses.

-Don


From: "Bob Camp" lists@rtty.us
Sent: Wednesday, January 09, 2013 6:43 AM
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] Interval Timer Recommendation

Hi

There are a number of counters that are a lot better than what you have.
There's also a limited amount of money in the universe. I'd suggest
setting a rational budget and time frame. This is all used gear. You
might get a perfect one that never ever breaks. More likely you will
need to get a couple and swap things around. I'd plan on buying 3 or 4
with what ever the budget is. If you get lucky, you come in under budget.

Less fancy:

5334 - couple of ns resolution, nice and small, B is very different than
the A so get one or the other
5335 - faster bigger heavier, one ns resolution.

both should be around $100 before shipping.

Very cool:

53131 500 ps resolution
53132 150 ps resolution

both have displays that can wear out. They are no longer supported by HP
so you might find somebody dumping them. Hard to find for less than
$800.

Big and heavy:

5371, 5372 same resolution as the 53132

They don't show up much. I've seen them for $200. You probably should
expect to pay $400. Expect shipping to be non-trivial.

High end:

SR620
5370
53230a

All are in the 20 ps range. You do see 5370's for $200. More commonly you
see them for quite a bit more. The 620's show up anywhere between $800 and
$3000. The 53230a is brand new with a warranty from Agilent for about
$3700.

That's just one set of parts. There are others. There are more issues than
just price to consider. The 5371 is a big piece of gear. The 53xxx boxes
are all small enough to carry around.

Lots of choices.

Bob

On Jan 8, 2013, at 9:50 PM, dlewis6767 dlewis6767@austin.rr.com wrote:

Hi Magnus, ....good day.

Thanks for the thoughts, ...that was pretty much where I was headed.  I
think I will hold out a while for a 5370, ....watch eBay, see if a
reasonably priced unit pops up.

If anyone has one for sale for a nice price, please let me know.

-Don


From: "Magnus Danielson" magnus@rubidium.dyndns.org
Sent: Tuesday, January 08, 2013 7:42 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Interval Timer Recommendation

Hi Don,

On 01/09/2013 02:32 AM, dlewis6767 wrote:

I am looking to upgrade my older HP 5245L counter/ timer to one with
greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers)
and just want to start measuring their 1PPS jitter, etc. over one
period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period
at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is
the best resolution they can provide for a one second pulse (over one
period)?

Do you have any recommendation as to which one of the three models is
the better unit?  (Their display technology, age and cost not
withstanding.) All three models seem greatly overstated in price on
eBay, seems, too. Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Of the given set, the HP5370 will have much better single shot
resolution compared to the HP5371/5372. The later is better at recording
high rate events.

I have both HP5370 and HP5372 as they fill different purposes. I also
have counters of better performance than these.

Trouble is, with PPS signals you experience jitter such that a 1 ns
resolution of a HP5335 will suffice for most usages.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Thank you Bob, Ed, Magnus , ..and all. Excellent reviews, ...detailed and clear. I REALLY do appreciate it. HP has turned out so much equipment over the years, it is often good to see a comparison. This was excellent. I'll wait for a good deal, ....and don the road bid alongside many of you , I presume. :-) I may opt for one of the better units, ...as this disease progresses. -Don -------------------------------------------------- From: "Bob Camp" <lists@rtty.us> Sent: Wednesday, January 09, 2013 6:43 AM To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Subject: Re: [time-nuts] Interval Timer Recommendation > Hi > > There are a number of counters that are a lot better than what you have. > There's also a limited amount of money in the universe. I'd suggest > setting a rational budget and time frame. This is all used gear. You > *might* get a perfect one that never ever breaks. More likely you will > need to get a couple and swap things around. I'd plan on buying 3 or 4 > with what ever the budget is. If you get lucky, you come in under budget. > > Less fancy: > > 5334 - couple of ns resolution, nice and small, B is *very* different than > the A so get one or the other > 5335 - faster bigger heavier, one ns resolution. > > both should be around $100 before shipping. > > Very cool: > > 53131 500 ps resolution > 53132 150 ps resolution > > both have displays that can wear out. They are no longer supported by HP > so you *might* find somebody dumping them. Hard to find for less than > $800. > > Big and heavy: > > 5371, 5372 same resolution as the 53132 > > They don't show up much. I've seen them for $200. You probably should > expect to pay $400. Expect shipping to be non-trivial. > > High end: > > SR620 > 5370 > 53230a > > All are in the 20 ps range. You do see 5370's for $200. More commonly you > see them for quite a bit more. The 620's show up anywhere between $800 and > $3000. The 53230a is brand new with a warranty from Agilent for about > $3700. > > That's just one set of parts. There are others. There are more issues than > just price to consider. The 5371 is a *big* piece of gear. The 53xxx boxes > are all small enough to carry around. > > Lots of choices. > > Bob > > On Jan 8, 2013, at 9:50 PM, dlewis6767 <dlewis6767@austin.rr.com> wrote: > >> Hi Magnus, ....good day. >> >> Thanks for the thoughts, ...that was pretty much where I was headed. I >> think I will hold out a while for a 5370, ....watch eBay, see if a >> reasonably priced unit pops up. >> >> If anyone has one for sale for a nice price, please let me know. >> >> -Don >> >> >> >> >> >> >> -------------------------------------------------- >> From: "Magnus Danielson" <magnus@rubidium.dyndns.org> >> Sent: Tuesday, January 08, 2013 7:42 PM >> To: <time-nuts@febo.com> >> Subject: Re: [time-nuts] Interval Timer Recommendation >> >>> Hi Don, >>> >>> On 01/09/2013 02:32 AM, dlewis6767 wrote: >>>> I am looking to upgrade my older HP 5245L counter/ timer to one with >>>> greater resolution. >>>> >>>> In my home lab, I have been collecting GPS timing modules (receivers) >>>> and just want to start measuring their 1PPS jitter, etc. over one >>>> period. >>>> >>>> The 5245L will only measure to 100ns with a 1 second pulse (.1us period >>>> at 1 period). >>>> >>>> I was looking at the HP5370/ 71/ 72 series counter/ timers. What is >>>> the best resolution they can provide for a one second pulse (over one >>>> period)? >>>> >>>> Do you have any recommendation as to which one of the three models is >>>> the better unit? (Their display technology, age and cost not >>>> withstanding.) All three models seem greatly overstated in price on >>>> eBay, seems, too. Not sure why, ...just supply and demand; I presume. >>>> >>>> Thank you for your advice; I appreciate your thoughts. >>> >>> Of the given set, the HP5370 will have much better single shot >>> resolution compared to the HP5371/5372. The later is better at recording >>> high rate events. >>> >>> I have both HP5370 and HP5372 as they fill different purposes. I also >>> have counters of better performance than these. >>> >>> Trouble is, with PPS signals you experience jitter such that a 1 ns >>> resolution of a HP5335 will suffice for most usages. >>> >>> Cheers, >>> Magnus >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Wed, Jan 9, 2013 1:28 PM

Hi

Keep in mind that the "good" prices are generally for gear that is in the
"don't know if it works at all" category. The prices go up quickly for stuff
that has had even basic checks run on it.
It's rare to find any of this stuff with a believable iron clad guarantee /
certification that it runs 100% to original specs. I've seen a lot of gear
with cal stickers on it that doesn't do this or that correctly...

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of dlewis6767
Sent: Wednesday, January 09, 2013 8:07 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Interval Timer Recommendation

Thank you Bob, Ed, Magnus , ..and all.

Excellent reviews, ...detailed and clear.  I REALLY do appreciate it.

HP has turned out so much equipment over the years, it is often good to see
a comparison.  This was excellent.

I'll wait for a good deal, ....and don the road bid alongside many of you ,
I presume.  :-)

I may opt for one of the better units, ...as this disease progresses.

-Don


From: "Bob Camp" lists@rtty.us
Sent: Wednesday, January 09, 2013 6:43 AM
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] Interval Timer Recommendation

Hi

There are a number of counters that are a lot better than what you have.
There's also a limited amount of money in the universe. I'd suggest
setting a rational budget and time frame. This is all used gear. You
might get a perfect one that never ever breaks. More likely you will
need to get a couple and swap things around. I'd plan on buying 3 or 4
with what ever the budget is. If you get lucky, you come in under budget.

Less fancy:

5334 - couple of ns resolution, nice and small, B is very different than

the A so get one or the other
5335 - faster bigger heavier, one ns resolution.

both should be around $100 before shipping.

Very cool:

53131 500 ps resolution
53132 150 ps resolution

both have displays that can wear out. They are no longer supported by HP
so you might find somebody dumping them. Hard to find for less than
$800.

Big and heavy:

5371, 5372 same resolution as the 53132

They don't show up much. I've seen them for $200. You probably should
expect to pay $400. Expect shipping to be non-trivial.

High end:

SR620
5370
53230a

All are in the 20 ps range. You do see 5370's for $200. More commonly you
see them for quite a bit more. The 620's show up anywhere between $800 and

$3000. The 53230a is brand new with a warranty from Agilent for about
$3700.

That's just one set of parts. There are others. There are more issues than

just price to consider. The 5371 is a big piece of gear. The 53xxx boxes

are all small enough to carry around.

Lots of choices.

Bob

On Jan 8, 2013, at 9:50 PM, dlewis6767 dlewis6767@austin.rr.com wrote:

Hi Magnus, ....good day.

Thanks for the thoughts, ...that was pretty much where I was headed.  I
think I will hold out a while for a 5370, ....watch eBay, see if a
reasonably priced unit pops up.

If anyone has one for sale for a nice price, please let me know.

-Don


From: "Magnus Danielson" magnus@rubidium.dyndns.org
Sent: Tuesday, January 08, 2013 7:42 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Interval Timer Recommendation

Hi Don,

On 01/09/2013 02:32 AM, dlewis6767 wrote:

I am looking to upgrade my older HP 5245L counter/ timer to one with
greater resolution.

In my home lab, I have been collecting GPS timing modules (receivers)
and just want to start measuring their 1PPS jitter, etc. over one
period.

The 5245L will only measure to 100ns with a 1 second pulse (.1us period

at 1 period).

I was looking at the HP5370/ 71/ 72 series counter/ timers.  What is
the best resolution they can provide for a one second pulse (over one
period)?

Do you have any recommendation as to which one of the three models is
the better unit?  (Their display technology, age and cost not
withstanding.) All three models seem greatly overstated in price on
eBay, seems, too. Not sure why, ...just supply and demand; I presume.

Thank you for your advice; I appreciate your thoughts.

Of the given set, the HP5370 will have much better single shot
resolution compared to the HP5371/5372. The later is better at recording

high rate events.

I have both HP5370 and HP5372 as they fill different purposes. I also
have counters of better performance than these.

Trouble is, with PPS signals you experience jitter such that a 1 ns
resolution of a HP5335 will suffice for most usages.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Keep in mind that the "good" prices are generally for gear that is in the "don't know if it works at all" category. The prices go up quickly for stuff that has had even basic checks run on it. It's rare to find any of this stuff with a believable iron clad guarantee / certification that it runs 100% to original specs. I've seen a lot of gear with cal stickers on it that doesn't do this or that correctly... Bob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of dlewis6767 Sent: Wednesday, January 09, 2013 8:07 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Interval Timer Recommendation Thank you Bob, Ed, Magnus , ..and all. Excellent reviews, ...detailed and clear. I REALLY do appreciate it. HP has turned out so much equipment over the years, it is often good to see a comparison. This was excellent. I'll wait for a good deal, ....and don the road bid alongside many of you , I presume. :-) I may opt for one of the better units, ...as this disease progresses. -Don -------------------------------------------------- From: "Bob Camp" <lists@rtty.us> Sent: Wednesday, January 09, 2013 6:43 AM To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Subject: Re: [time-nuts] Interval Timer Recommendation > Hi > > There are a number of counters that are a lot better than what you have. > There's also a limited amount of money in the universe. I'd suggest > setting a rational budget and time frame. This is all used gear. You > *might* get a perfect one that never ever breaks. More likely you will > need to get a couple and swap things around. I'd plan on buying 3 or 4 > with what ever the budget is. If you get lucky, you come in under budget. > > Less fancy: > > 5334 - couple of ns resolution, nice and small, B is *very* different than > the A so get one or the other > 5335 - faster bigger heavier, one ns resolution. > > both should be around $100 before shipping. > > Very cool: > > 53131 500 ps resolution > 53132 150 ps resolution > > both have displays that can wear out. They are no longer supported by HP > so you *might* find somebody dumping them. Hard to find for less than > $800. > > Big and heavy: > > 5371, 5372 same resolution as the 53132 > > They don't show up much. I've seen them for $200. You probably should > expect to pay $400. Expect shipping to be non-trivial. > > High end: > > SR620 > 5370 > 53230a > > All are in the 20 ps range. You do see 5370's for $200. More commonly you > see them for quite a bit more. The 620's show up anywhere between $800 and > $3000. The 53230a is brand new with a warranty from Agilent for about > $3700. > > That's just one set of parts. There are others. There are more issues than > just price to consider. The 5371 is a *big* piece of gear. The 53xxx boxes > are all small enough to carry around. > > Lots of choices. > > Bob > > On Jan 8, 2013, at 9:50 PM, dlewis6767 <dlewis6767@austin.rr.com> wrote: > >> Hi Magnus, ....good day. >> >> Thanks for the thoughts, ...that was pretty much where I was headed. I >> think I will hold out a while for a 5370, ....watch eBay, see if a >> reasonably priced unit pops up. >> >> If anyone has one for sale for a nice price, please let me know. >> >> -Don >> >> >> >> >> >> >> -------------------------------------------------- >> From: "Magnus Danielson" <magnus@rubidium.dyndns.org> >> Sent: Tuesday, January 08, 2013 7:42 PM >> To: <time-nuts@febo.com> >> Subject: Re: [time-nuts] Interval Timer Recommendation >> >>> Hi Don, >>> >>> On 01/09/2013 02:32 AM, dlewis6767 wrote: >>>> I am looking to upgrade my older HP 5245L counter/ timer to one with >>>> greater resolution. >>>> >>>> In my home lab, I have been collecting GPS timing modules (receivers) >>>> and just want to start measuring their 1PPS jitter, etc. over one >>>> period. >>>> >>>> The 5245L will only measure to 100ns with a 1 second pulse (.1us period >>>> at 1 period). >>>> >>>> I was looking at the HP5370/ 71/ 72 series counter/ timers. What is >>>> the best resolution they can provide for a one second pulse (over one >>>> period)? >>>> >>>> Do you have any recommendation as to which one of the three models is >>>> the better unit? (Their display technology, age and cost not >>>> withstanding.) All three models seem greatly overstated in price on >>>> eBay, seems, too. Not sure why, ...just supply and demand; I presume. >>>> >>>> Thank you for your advice; I appreciate your thoughts. >>> >>> Of the given set, the HP5370 will have much better single shot >>> resolution compared to the HP5371/5372. The later is better at recording >>> high rate events. >>> >>> I have both HP5370 and HP5372 as they fill different purposes. I also >>> have counters of better performance than these. >>> >>> Trouble is, with PPS signals you experience jitter such that a 1 ns >>> resolution of a HP5335 will suffice for most usages. >>> >>> Cheers, >>> Magnus >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
MD
Magnus Danielson
Wed, Jan 9, 2013 6:29 PM

On 01/09/2013 05:39 AM, Ed Palmer wrote:

In the past there has been much discussion on this list regarding the
relative strengths and weaknesses of the 5370 (20ps resolution) and the
5371/5372 (150ps resolution) units. There are other units that have even
better resolution, but they tend to be pricey and/or more specialized.

The 5371/5372 has a 200 ps resolution. They where designed to allow
extension to 100 ps resolution. The FLASH-interpolators have 200 ps
steps, sufficiently good trigger such that you get 150 ps resolution
performance on the white-noise limit.

Cheers,
Magnus

On 01/09/2013 05:39 AM, Ed Palmer wrote: > In the past there has been much discussion on this list regarding the > relative strengths and weaknesses of the 5370 (20ps resolution) and the > 5371/5372 (150ps resolution) units. There are other units that have even > better resolution, but they tend to be pricey and/or more specialized. The 5371/5372 has a 200 ps resolution. They where designed to allow extension to 100 ps resolution. The FLASH-interpolators have 200 ps steps, sufficiently good trigger such that you get 150 ps resolution performance on the white-noise limit. Cheers, Magnus
AB
Azelio Boriani
Wed, Jan 9, 2013 8:48 PM

and don't forget the PM6681 (50pS)

On Wed, Jan 9, 2013 at 7:29 PM, Magnus Danielson <magnus@rubidium.dyndns.org

wrote:

On 01/09/2013 05:39 AM, Ed Palmer wrote:

In the past there has been much discussion on this list regarding the
relative strengths and weaknesses of the 5370 (20ps resolution) and the
5371/5372 (150ps resolution) units. There are other units that have even
better resolution, but they tend to be pricey and/or more specialized.

The 5371/5372 has a 200 ps resolution. They where designed to allow
extension to 100 ps resolution. The FLASH-interpolators have 200 ps steps,
sufficiently good trigger such that you get 150 ps resolution performance
on the white-noise limit.

Cheers,
Magnus

_____________**
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**
mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

and don't forget the PM6681 (50pS) On Wed, Jan 9, 2013 at 7:29 PM, Magnus Danielson <magnus@rubidium.dyndns.org > wrote: > On 01/09/2013 05:39 AM, Ed Palmer wrote: > >> In the past there has been much discussion on this list regarding the >> relative strengths and weaknesses of the 5370 (20ps resolution) and the >> 5371/5372 (150ps resolution) units. There are other units that have even >> better resolution, but they tend to be pricey and/or more specialized. >> > > The 5371/5372 has a 200 ps resolution. They where designed to allow > extension to 100 ps resolution. The FLASH-interpolators have 200 ps steps, > sufficiently good trigger such that you get 150 ps resolution performance > on the white-noise limit. > > > Cheers, > Magnus > > ______________________________**_________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/time-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. >
R
Rex
Wed, Jan 9, 2013 11:02 PM

On 1/9/2013 12:48 PM, Azelio Boriani wrote:

and don't forget the PM6681 (50pS)

The PM6681 was sold by Fluke/Philips. The same counter is also
occasionally seen as the Pendulum CNT-81. Additional good features:
small, light, and quiet.

On Wed, Jan 9, 2013 at 7:29 PM, Magnus Danielson<magnus@rubidium.dyndns.org

wrote:
On 01/09/2013 05:39 AM, Ed Palmer wrote:

In the past there has been much discussion on this list regarding the
relative strengths and weaknesses of the 5370 (20ps resolution) and the
5371/5372 (150ps resolution) units. There are other units that have even
better resolution, but they tend to be pricey and/or more specialized.

The 5371/5372 has a 200 ps resolution. They where designed to allow
extension to 100 ps resolution. The FLASH-interpolators have 200 ps steps,
sufficiently good trigger such that you get 150 ps resolution performance
on the white-noise limit.

Cheers,
Magnus

On 1/9/2013 12:48 PM, Azelio Boriani wrote: > and don't forget the PM6681 (50pS) The PM6681 was sold by Fluke/Philips. The same counter is also occasionally seen as the Pendulum CNT-81. Additional good features: small, light, and quiet. > On Wed, Jan 9, 2013 at 7:29 PM, Magnus Danielson<magnus@rubidium.dyndns.org >> wrote: >> On 01/09/2013 05:39 AM, Ed Palmer wrote: >> >>> In the past there has been much discussion on this list regarding the >>> relative strengths and weaknesses of the 5370 (20ps resolution) and the >>> 5371/5372 (150ps resolution) units. There are other units that have even >>> better resolution, but they tend to be pricey and/or more specialized. >>> >> The 5371/5372 has a 200 ps resolution. They where designed to allow >> extension to 100 ps resolution. The FLASH-interpolators have 200 ps steps, >> sufficiently good trigger such that you get 150 ps resolution performance >> on the white-noise limit. >> >> >> Cheers, >> Magnus >>
MD
Magnus Danielson
Thu, Jan 10, 2013 3:39 AM

On 01/10/2013 12:02 AM, Rex wrote:

On 1/9/2013 12:48 PM, Azelio Boriani wrote:

and don't forget the PM6681 (50pS)

The PM6681 was sold by Fluke/Philips. The same counter is also
occasionally seen as the Pendulum CNT-81. Additional good features:
small, light, and quiet.

A bit more history then.

Philips had a sub-sidary called Philips Industrier Järfälla, located
just outside Stockholm. They made amongst other things OCXOs and
counters, but a numerous of other measuring gadgets. As Philips merged
with Fluke for instruments, the Philips Industrier measurments gear also
re-occured in their Fluke variant for the US market. They kept their
PMxxxx model numbers. As Philips was retracting from the market, they
sold of the T&F reminder of Philips Industrier and that was named
Pendulum, which kept the connection with Fluke. This meant that the new
counters coming out also came out in their Fluke variants. However,
pendulum decided to strike out the PM66 prefix (for counters) for their
Pendulum branded gear and had different prefixes such as
CNT-80/81/85/90/90XL/91, GPS-12/88/89 and WM-10/11. Pendulum was sold
off to SpectraCom, and it was run as a separate business for some time,
but eventually the old company was dismantled and Pendulum remains as a
brand within SpectraCom. The product range exists, except for the CNT-8x
series which died off then the last of the custom counter chips ran out,
besides what was kept for service of course. The CNT-9x series uses a
FPGA as counter core. Today only a few of the original staff remains
with SpectraCom. The CNT-90 has also appeared in Tektronix format.

I had some exchange with them as they where local to me.

Cheers,
Magnus

On 01/10/2013 12:02 AM, Rex wrote: > On 1/9/2013 12:48 PM, Azelio Boriani wrote: >> and don't forget the PM6681 (50pS) > > The PM6681 was sold by Fluke/Philips. The same counter is also > occasionally seen as the Pendulum CNT-81. Additional good features: > small, light, and quiet. A bit more history then. Philips had a sub-sidary called Philips Industrier Järfälla, located just outside Stockholm. They made amongst other things OCXOs and counters, but a numerous of other measuring gadgets. As Philips merged with Fluke for instruments, the Philips Industrier measurments gear also re-occured in their Fluke variant for the US market. They kept their PMxxxx model numbers. As Philips was retracting from the market, they sold of the T&F reminder of Philips Industrier and that was named Pendulum, which kept the connection with Fluke. This meant that the new counters coming out also came out in their Fluke variants. However, pendulum decided to strike out the PM66 prefix (for counters) for their Pendulum branded gear and had different prefixes such as CNT-80/81/85/90/90XL/91, GPS-12/88/89 and WM-10/11. Pendulum was sold off to SpectraCom, and it was run as a separate business for some time, but eventually the old company was dismantled and Pendulum remains as a brand within SpectraCom. The product range exists, except for the CNT-8x series which died off then the last of the custom counter chips ran out, besides what was kept for service of course. The CNT-9x series uses a FPGA as counter core. Today only a few of the original staff remains with SpectraCom. The CNT-90 has also appeared in Tektronix format. I had some exchange with them as they where local to me. Cheers, Magnus
C
cfo
Thu, Jan 10, 2013 5:37 PM

On Wed, 09 Jan 2013 21:48:00 +0100, Azelio Boriani wrote:

and don't forget the PM6681 (50pS)

I have a PM6680 (250 or 300pS) i think , but it was within my budget.

I just read the Service manual , and noticed that both the PM6680 & the
6681 has a battery inside that holds the settings.
The manual states that loosing the settings on a 6680 is no big problem ,
but on the 6681 you loose the interpolator calibtation. And it sounds
like that's not a good thing.
I didn't know you had to watch out for Battery on a PM6680/81.

CFO - Tnut-Beginner
Denmark

On Wed, 09 Jan 2013 21:48:00 +0100, Azelio Boriani wrote: > and don't forget the PM6681 (50pS) > I have a PM6680 (250 or 300pS) i think , but it was within my budget. I just read the Service manual , and noticed that both the PM6680 & the 6681 has a battery inside that holds the settings. The manual states that loosing the settings on a 6680 is no big problem , but on the 6681 you loose the interpolator calibtation. And it sounds like that's not a good thing. I didn't know you had to watch out for Battery on a PM6680/81. CFO - Tnut-Beginner Denmark
AB
Azelio Boriani
Thu, Jan 10, 2013 10:14 PM

Yes, the battery on the PM6681 is a real problem, unless you have the
unobtainable calibration program. The program uses the GPIB to perform the
calibration: I have tried to figure out the GPIB commands involved in the
calibration process by looking at the firmware dump... the commands are
clear text. I wasn't able to find anything.

On Thu, Jan 10, 2013 at 6:37 PM, cfo xnews3@luna.dyndns.dk wrote:

On Wed, 09 Jan 2013 21:48:00 +0100, Azelio Boriani wrote:

and don't forget the PM6681 (50pS)

I have a PM6680 (250 or 300pS) i think , but it was within my budget.

I just read the Service manual , and noticed that both the PM6680 & the
6681 has a battery inside that holds the settings.
The manual states that loosing the settings on a 6680 is no big problem ,
but on the 6681 you loose the interpolator calibtation. And it sounds
like that's not a good thing.
I didn't know you had to watch out for Battery on a PM6680/81.

CFO - Tnut-Beginner
Denmark


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Yes, the battery on the PM6681 is a real problem, unless you have the unobtainable calibration program. The program uses the GPIB to perform the calibration: I have tried to figure out the GPIB commands involved in the calibration process by looking at the firmware dump... the commands are clear text. I wasn't able to find anything. On Thu, Jan 10, 2013 at 6:37 PM, cfo <xnews3@luna.dyndns.dk> wrote: > On Wed, 09 Jan 2013 21:48:00 +0100, Azelio Boriani wrote: > > > and don't forget the PM6681 (50pS) > > > I have a PM6680 (250 or 300pS) i think , but it was within my budget. > > I just read the Service manual , and noticed that both the PM6680 & the > 6681 has a battery inside that holds the settings. > The manual states that loosing the settings on a 6680 is no big problem , > but on the 6681 you loose the interpolator calibtation. And it sounds > like that's not a good thing. > I didn't know you had to watch out for Battery on a PM6680/81. > > CFO - Tnut-Beginner > Denmark > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
JW
Jonatan Walck
Fri, Jan 11, 2013 12:03 AM

On 2013-01-10 23:14, Azelio Boriani wrote:

Yes, the battery on the PM6681 is a real problem, unless you have
the unobtainable calibration program. The program uses the GPIB to
perform the calibration: I have tried to figure out the GPIB
commands involved in the calibration process by looking at the
firmware dump... the commands are clear text. I wasn't able to find
anything.

I did read about the battery before, but did not know how severe the
problem was. Thanks for the heads up! Does the PM6685 end up in the same
category as 80 or 81 in this respect?

I want to send a thank you to everyone participating on this thread, I'm
sure I'm not alone looking for a (or another) time-nut friendly TIC.
Finding one that's, "good enough" electrially (all parameters up to
desired specs), user friendly, not humongous and affordable was harder
than I initially thought it would be.

So far I've seen the Racal-Dana 1991/1992, horde of HP/Agilents and
Philips/Fluke/Pendulum CNT-/PM66* mentioned here. Keeping a searching
eye out for one that isn't 19" and doesn't weight too much given the
frequency of which I will lug it around. Or failing that cheap enough
to double up.:)

To me this so far means I'm primairly looking for Racal-Dana's as a
cheap option or possibly PM66's if I can find them (much rarer here, but
I like that they're made in Sweden!:). Are there any HP's I should keep
a look out for, if I'm aiming for at least 1 ns resolution? Bob's list
(a special thanks to you for the comprehensive list) left me with only
finding either very expensive or very big ones in Europe.

Any recommendations for where to look for timing equipment (TICs
specifically but I guess it will be the same answer in general) in
Europe besides the one all so familiar auction site? (Not that I have
anything against ebay, just checking if there are any secondary sources.)

// jwalck

On 2013-01-10 23:14, Azelio Boriani wrote: > Yes, the battery on the PM6681 is a real problem, unless you have > the unobtainable calibration program. The program uses the GPIB to > perform the calibration: I have tried to figure out the GPIB > commands involved in the calibration process by looking at the > firmware dump... the commands are clear text. I wasn't able to find > anything. I did read about the battery before, but did not know how severe the problem was. Thanks for the heads up! Does the PM6685 end up in the same category as 80 or 81 in this respect? I want to send a thank you to everyone participating on this thread, I'm sure I'm not alone looking for a (or another) time-nut friendly TIC. Finding one that's, "good enough" electrially (all parameters up to desired specs), user friendly, not humongous and affordable was harder than I initially thought it would be. So far I've seen the Racal-Dana 1991/1992, horde of HP/Agilents and Philips/Fluke/Pendulum CNT-/PM66* mentioned here. Keeping a searching eye out for one that isn't 19" and doesn't weight too much given the frequency of which I will lug it around. Or failing that cheap enough to double up.:) To me this so far means I'm primairly looking for Racal-Dana's as a cheap option or possibly PM66's if I can find them (much rarer here, but I like that they're made in Sweden!:). Are there any HP's I should keep a look out for, if I'm aiming for at least 1 ns resolution? Bob's list (a special thanks to you for the comprehensive list) left me with only finding either very expensive or very big ones in Europe. Any recommendations for where to look for timing equipment (TICs specifically but I guess it will be the same answer in general) in Europe besides the one all so familiar auction site? (Not that I have anything against ebay, just checking if there are any secondary sources.) // jwalck
BC
Bob Camp
Fri, Jan 11, 2013 1:48 AM

HI

One thing missing from the list was the HP 53181. It's useless as a timing device and thus didn't get listed. If you are after frequency only, they occasionally show up for quite a bit less than the 53131's. As a frequency counter, they have the same performance as a '131. All of the 53131, 132, 181, and 230's are small enough to lug around.

Bob

On Jan 10, 2013, at 7:03 PM, Jonatan Walck jwalck@netnod.se wrote:

On 2013-01-10 23:14, Azelio Boriani wrote:

Yes, the battery on the PM6681 is a real problem, unless you have
the unobtainable calibration program. The program uses the GPIB to
perform the calibration: I have tried to figure out the GPIB
commands involved in the calibration process by looking at the
firmware dump... the commands are clear text. I wasn't able to find
anything.

I did read about the battery before, but did not know how severe the
problem was. Thanks for the heads up! Does the PM6685 end up in the same
category as 80 or 81 in this respect?

I want to send a thank you to everyone participating on this thread, I'm
sure I'm not alone looking for a (or another) time-nut friendly TIC.
Finding one that's, "good enough" electrially (all parameters up to
desired specs), user friendly, not humongous and affordable was harder
than I initially thought it would be.

So far I've seen the Racal-Dana 1991/1992, horde of HP/Agilents and
Philips/Fluke/Pendulum CNT-/PM66* mentioned here. Keeping a searching
eye out for one that isn't 19" and doesn't weight too much given the
frequency of which I will lug it around. Or failing that cheap enough
to double up.:)

To me this so far means I'm primairly looking for Racal-Dana's as a
cheap option or possibly PM66's if I can find them (much rarer here, but
I like that they're made in Sweden!:). Are there any HP's I should keep
a look out for, if I'm aiming for at least 1 ns resolution? Bob's list
(a special thanks to you for the comprehensive list) left me with only
finding either very expensive or very big ones in Europe.

Any recommendations for where to look for timing equipment (TICs
specifically but I guess it will be the same answer in general) in
Europe besides the one all so familiar auction site? (Not that I have
anything against ebay, just checking if there are any secondary sources.)

// jwalck


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

HI One thing missing from the list was the HP 53181. It's useless as a timing device and thus didn't get listed. If you are after frequency only, they occasionally show up for quite a bit less than the 53131's. As a frequency counter, they have the same performance as a '131. All of the 53131, 132, 181, and 230's are small enough to lug around. Bob On Jan 10, 2013, at 7:03 PM, Jonatan Walck <jwalck@netnod.se> wrote: > On 2013-01-10 23:14, Azelio Boriani wrote: >> Yes, the battery on the PM6681 is a real problem, unless you have >> the unobtainable calibration program. The program uses the GPIB to >> perform the calibration: I have tried to figure out the GPIB >> commands involved in the calibration process by looking at the >> firmware dump... the commands are clear text. I wasn't able to find >> anything. > > I did read about the battery before, but did not know how severe the > problem was. Thanks for the heads up! Does the PM6685 end up in the same > category as 80 or 81 in this respect? > > > I want to send a thank you to everyone participating on this thread, I'm > sure I'm not alone looking for a (or another) time-nut friendly TIC. > Finding one that's, "good enough" electrially (all parameters up to > desired specs), user friendly, not humongous and affordable was harder > than I initially thought it would be. > > So far I've seen the Racal-Dana 1991/1992, horde of HP/Agilents and > Philips/Fluke/Pendulum CNT-/PM66* mentioned here. Keeping a searching > eye out for one that isn't 19" and doesn't weight too much given the > frequency of which I will lug it around. Or failing that cheap enough > to double up.:) > > To me this so far means I'm primairly looking for Racal-Dana's as a > cheap option or possibly PM66's if I can find them (much rarer here, but > I like that they're made in Sweden!:). Are there any HP's I should keep > a look out for, if I'm aiming for at least 1 ns resolution? Bob's list > (a special thanks to you for the comprehensive list) left me with only > finding either very expensive or very big ones in Europe. > > > Any recommendations for where to look for timing equipment (TICs > specifically but I guess it will be the same answer in general) in > Europe besides the one all so familiar auction site? (Not that I have > anything against ebay, just checking if there are any secondary sources.) > > // jwalck > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
CH
Chris Howard
Fri, Jan 11, 2013 2:51 PM

The manual states that loosing the settings on a 6680 is no big problem ,
but on the 6681 you loose the interpolator calibtation. And it sounds
like that's not a good thing.
I didn't know you had to watch out for Battery on a PM6680/81.

CFO - Tnut-Beginner
Denmark

Would loss of interpolation calibration have much impact on its use as a TIC?

>> The manual states that loosing the settings on a 6680 is no big problem , >> but on the 6681 you loose the interpolator calibtation. And it sounds >> like that's not a good thing. >> I didn't know you had to watch out for Battery on a PM6680/81. >> >> CFO - Tnut-Beginner >> Denmark Would loss of interpolation calibration have much impact on its use as a TIC?
MD
Magnus Danielson
Fri, Jan 11, 2013 6:33 PM

On 01/11/2013 03:51 PM, Chris Howard wrote:

The manual states that loosing the settings on a 6680 is no big problem ,
but on the 6681 you loose the interpolator calibtation. And it sounds
like that's not a good thing.
I didn't know you had to watch out for Battery on a PM6680/81.

CFO - Tnut-Beginner
Denmark

Would loss of interpolation calibration have much impact on its use as a TIC?

Yes, as that is what gives you the precision. I don't recall how bad it
will get, but I ended up having the PM6681 calibrated.

Cheers,
Magnus

On 01/11/2013 03:51 PM, Chris Howard wrote: > >>> The manual states that loosing the settings on a 6680 is no big problem , >>> but on the 6681 you loose the interpolator calibtation. And it sounds >>> like that's not a good thing. >>> I didn't know you had to watch out for Battery on a PM6680/81. >>> >>> CFO - Tnut-Beginner >>> Denmark > > > Would loss of interpolation calibration have much impact on its use as a TIC? Yes, as that is what gives you the precision. I don't recall how bad it will get, but I ended up having the PM6681 calibrated. Cheers, Magnus
AB
Azelio Boriani
Fri, Jan 11, 2013 8:09 PM

The PM6685 is a frequency counter, not a TIC, so it shouldn't suffer from
the battery problem.

On Fri, Jan 11, 2013 at 7:33 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:

On 01/11/2013 03:51 PM, Chris Howard wrote:

The manual states that loosing the settings on a 6680 is no big problem ,

but on the 6681 you loose the interpolator calibtation. And it sounds
like that's not a good thing.
I didn't know you had to watch out for Battery on a PM6680/81.

CFO - Tnut-Beginner
Denmark

Would loss of interpolation calibration have much impact on its use as a
TIC?

Yes, as that is what gives you the precision. I don't recall how bad it
will get, but I ended up having the PM6681 calibrated.

Cheers,
Magnus

_____________**
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**
mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

The PM6685 is a frequency counter, not a TIC, so it shouldn't suffer from the battery problem. On Fri, Jan 11, 2013 at 7:33 PM, Magnus Danielson < magnus@rubidium.dyndns.org> wrote: > On 01/11/2013 03:51 PM, Chris Howard wrote: > >> >> The manual states that loosing the settings on a 6680 is no big problem , >>>> but on the 6681 you loose the interpolator calibtation. And it sounds >>>> like that's not a good thing. >>>> I didn't know you had to watch out for Battery on a PM6680/81. >>>> >>>> CFO - Tnut-Beginner >>>> Denmark >>>> >>> >> >> Would loss of interpolation calibration have much impact on its use as a >> TIC? >> > > Yes, as that is what gives you the precision. I don't recall how bad it > will get, but I ended up having the PM6681 calibrated. > > Cheers, > Magnus > > > ______________________________**_________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/time-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. >
MD
Magnus Danielson
Fri, Jan 11, 2013 9:15 PM

On 01/11/2013 09:09 PM, Azelio Boriani wrote:

The PM6685 is a frequency counter, not a TIC, so it shouldn't suffer from
the battery problem.

It's single shot resolution would still be from the interpolator, even
if you measure on the same signal.

Cheers,
Magnus

On 01/11/2013 09:09 PM, Azelio Boriani wrote: > The PM6685 is a frequency counter, not a TIC, so it shouldn't suffer from > the battery problem. It's single shot resolution would still be from the interpolator, even if you measure on the same signal. Cheers, Magnus
AB
Azelio Boriani
Sat, Jan 12, 2013 12:37 PM

OK, but I wasn't able to find a single shot resolution specification in its
datasheet. Looking at the formulas to determine the accuracy, it looks like
the minimum resolution is 250pS... anyway, by reading the user and service
manual I found out that the PM6685 has an EEPROM memory and no battery is
needed.

On Fri, Jan 11, 2013 at 10:15 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:

On 01/11/2013 09:09 PM, Azelio Boriani wrote:

The PM6685 is a frequency counter, not a TIC, so it shouldn't suffer from
the battery problem.

It's single shot resolution would still be from the interpolator, even if
you measure on the same signal.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

OK, but I wasn't able to find a single shot resolution specification in its datasheet. Looking at the formulas to determine the accuracy, it looks like the minimum resolution is 250pS... anyway, by reading the user and service manual I found out that the PM6685 has an EEPROM memory and no battery is needed. On Fri, Jan 11, 2013 at 10:15 PM, Magnus Danielson < magnus@rubidium.dyndns.org> wrote: > On 01/11/2013 09:09 PM, Azelio Boriani wrote: > >> The PM6685 is a frequency counter, not a TIC, so it shouldn't suffer from >> the battery problem. >> > > It's single shot resolution would still be from the interpolator, even if > you measure on the same signal. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Sat, Jan 12, 2013 1:40 PM

On 01/12/2013 01:37 PM, Azelio Boriani wrote:

OK, but I wasn't able to find a single shot resolution specification in its
datasheet. Looking at the formulas to determine the accuracy, it looks like
the minimum resolution is 250pS... anyway, by reading the user and service
manual I found out that the PM6685 has an EEPROM memory and no battery is
needed.

At 250 ps you use interpolators. The typical coarse counter would be 100
MHz and that means 10 ns in raw resolution, but you would then use an
interpolator to get higher resolution. By getting a start and stop
measure, roughly tau time inbetween each other you get
f = (t_stop - t_start)/tau
assuming t_start and t_stop trigger from the same source.

For frequency measures you can overcome the single-shot resolution by
using multiple start and stop locations and process the estimates.

BTW, some counters use quite different approaches for triggering and
routing their TI measures and their frequency measures, to such degree
that frequency trimming requires separate calibration. SR620 is one such
example. Typically would the calibration error cause a static time error
between start and stop, and scaling with tau is a good way to track it.
Measuring it's own time-base is a good way to measure this calibration
error and trim it out.

Cheers,
Magnus

On 01/12/2013 01:37 PM, Azelio Boriani wrote: > OK, but I wasn't able to find a single shot resolution specification in its > datasheet. Looking at the formulas to determine the accuracy, it looks like > the minimum resolution is 250pS... anyway, by reading the user and service > manual I found out that the PM6685 has an EEPROM memory and no battery is > needed. At 250 ps you use interpolators. The typical coarse counter would be 100 MHz and that means 10 ns in raw resolution, but you would then use an interpolator to get higher resolution. By getting a start and stop measure, roughly tau time inbetween each other you get f = (t_stop - t_start)/tau assuming t_start and t_stop trigger from the same source. For frequency measures you can overcome the single-shot resolution by using multiple start and stop locations and process the estimates. BTW, some counters use quite different approaches for triggering and routing their TI measures and their frequency measures, to such degree that frequency trimming requires separate calibration. SR620 is one such example. Typically would the calibration error cause a static time error between start and stop, and scaling with tau is a good way to track it. Measuring it's own time-base is a good way to measure this calibration error and trim it out. Cheers, Magnus
CH
Chris Howard
Sat, Jan 12, 2013 6:03 PM

On 1/12/2013 7:40 AM, Magnus Danielson wrote:

At 250 ps you use interpolators. The typical coarse counter would be 100 MHz and that means 10 ns in raw resolution, but you would then use an interpolator to get higher resolution. By getting a start and stop measure, roughly tau time inbetween each other you get
f = (t_stop - t_start)/tau
assuming t_start and t_stop trigger from the same source.

This may not be of interest, and if so, I'll drop it.
But I was thinking that if they go to some kind of
fallback interpolation the time interval measurements
at a more-or-less fixed frequency might still be
pretty good.

(I think along these lines because maybe someday
a battery-bad unit might show up which is not so good
at umpty-GHz measurements but still OK for TIC)

If it had a systematic error at 10MHz it seems
like maybe that would just fall out in the wash. (?)

Chris

On 1/12/2013 7:40 AM, Magnus Danielson wrote: > > At 250 ps you use interpolators. The typical coarse counter would be 100 MHz and that means 10 ns in raw resolution, but you would then use an interpolator to get higher resolution. By getting a start and stop measure, roughly tau time inbetween each other you get > f = (t_stop - t_start)/tau > assuming t_start and t_stop trigger from the same source. > This may not be of interest, and if so, I'll drop it. But I was thinking that if they go to some kind of fallback interpolation the time interval measurements at a more-or-less fixed frequency might still be pretty good. (I think along these lines because maybe someday a battery-bad unit might show up which is not so good at umpty-GHz measurements but still OK for TIC) If it had a systematic error at 10MHz it seems like maybe that would just fall out in the wash. (?) Chris
JW
Jonatan Walck
Mon, Jan 14, 2013 9:04 AM

-----BEGIN PGP SIGNED MESSAGE-----
Hash: SHA1

On 2013-01-11 02:48, Bob Camp wrote:

HI

One thing missing from the list was the HP 53181. It's useless as
a timing device and thus didn't get listed. If you are after
frequency only, they occasionally show up for quite a bit less than
the 53131's. As a frequency counter, they have the same performance
as a '131. All of the 53131, 132, 181, and 230's are small enough
to lug around.

Bob

I'll find uses for a full TIC rather than just a fequency counter, but
I absorb all this knowledge happily either way and it's good to have
that here as a reference.

I initially found the 53131A and everything above it to steep for my
first home lab TIC but you are right about the size, they are quite
small. (Having access to 53131's through work.)

As great luck would have it though, I found a second hand Pendulum
CNT-90 and I got a too good deal to hold back my trigger finger.
Picked it up yesterday! Made no measurements past "does it start up
and can it detect any frequency?" yet but will put it to good use soon.

I wouldn't have know what to look for and find this TIC without this
list. A big thanks to all time-nuts once again!

// jwalck
-----BEGIN PGP SIGNATURE-----
Version: GnuPG v1.4.12 (GNU/Linux)
Comment: Using GnuPG with Mozilla - http://www.enigmail.net/
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=lCIv
-----END PGP SIGNATURE-----

-----BEGIN PGP SIGNED MESSAGE----- Hash: SHA1 On 2013-01-11 02:48, Bob Camp wrote: > HI > > One thing missing from the list was the HP 53181. It's useless as > a timing device and thus didn't get listed. If you are after > frequency only, they occasionally show up for quite a bit less than > the 53131's. As a frequency counter, they have the same performance > as a '131. All of the 53131, 132, 181, and 230's are small enough > to lug around. > > Bob I'll find uses for a full TIC rather than just a fequency counter, but I absorb all this knowledge happily either way and it's good to have that here as a reference. I initially found the 53131A and everything above it to steep for my first home lab TIC but you are right about the size, they are quite small. (Having access to 53131's through work.) As great luck would have it though, I found a second hand Pendulum CNT-90 and I got a too good deal to hold back my trigger finger. Picked it up yesterday! Made no measurements past "does it start up and can it detect any frequency?" yet but will put it to good use soon. I wouldn't have know what to look for and find this TIC without this list. A big thanks to all time-nuts once again! // jwalck -----BEGIN PGP SIGNATURE----- Version: GnuPG v1.4.12 (GNU/Linux) Comment: Using GnuPG with Mozilla - http://www.enigmail.net/ iQIcBAEBAgAGBQJQ88oBAAoJEFwg9i9GDX+n+mMP/iWYkLQd4o3zVglTLESM8Gqu yo9SGThQQ1Uc4svaNLzHP6WC6uO5Dh+KG6oUDhDeeqL4CNDoCCs2F4lbMsavxgW/ lcrDjSCqATgB7J8edlLP/lw2q+Uo4VY6Y1qFmAah+6ooqU5D+mdi5j7Solm4SF+T O+VKgghklgPuw1mWWaESNgKvv869nNkMTKw/hwQF4DDiCu/9N/gv4bxt5P778oWv r8nu4Yck4AWWhGkDn3JcpCXqIGrA2T40CR5ek1TD9i6LGLBnCU7FmIrofd/fZEF3 AJVYfB+tIL6CO5jj+Xvk4G4HeOXMhknKZBVmF/kFO27q4IKFAbYhIOAWdWK1MwQT tO7N4banehXrkFuDJwJdZuze+OweME69/uBv2Rcj6jC4IWDk0kmcJKrcbGBzwuN9 S01jyea9Ndi1CVXfdYkY9C9Pp1FZM9FzIleSB7+IGYfgS6vU+lXNnq8OVndYiT5J 6zNjpOjd1NUWCU7WCqxLQmMEGNw1CG7ITY8G8gPAAsCsg2AkYKHix8WYidAvlc7o NBkp7kFhD4l2jD9N8Uu50+Y2/meliiJK2wFgwbTdLOKNfdVjQYwsJ7QcQ8vJtHju oxsESC/FVIvU0CyJ81tl7xQPV1kG9XQFuMKEISIyqCsA84+kApFP++sguM8RRLJ7 ctm8Is0scpOVQ7kLkkmm =lCIv -----END PGP SIGNATURE-----
PL
Pete Lancashire
Tue, Jan 15, 2013 2:34 PM

A bit late in the thread but I don't see the HP E1740A

-pete

On 1/14/13, Jonatan Walck jwalck@netnod.se wrote:

-----BEGIN PGP SIGNED MESSAGE-----
Hash: SHA1

On 2013-01-11 02:48, Bob Camp wrote:

HI

One thing missing from the list was the HP 53181. It's useless as
a timing device and thus didn't get listed. If you are after
frequency only, they occasionally show up for quite a bit less than
the 53131's. As a frequency counter, they have the same performance
as a '131. All of the 53131, 132, 181, and 230's are small enough
to lug around.

Bob

I'll find uses for a full TIC rather than just a fequency counter, but
I absorb all this knowledge happily either way and it's good to have
that here as a reference.

I initially found the 53131A and everything above it to steep for my
first home lab TIC but you are right about the size, they are quite
small. (Having access to 53131's through work.)

As great luck would have it though, I found a second hand Pendulum
CNT-90 and I got a too good deal to hold back my trigger finger.
Picked it up yesterday! Made no measurements past "does it start up
and can it detect any frequency?" yet but will put it to good use soon.

I wouldn't have know what to look for and find this TIC without this
list. A big thanks to all time-nuts once again!

// jwalck
-----BEGIN PGP SIGNATURE-----
Version: GnuPG v1.4.12 (GNU/Linux)
Comment: Using GnuPG with Mozilla - http://www.enigmail.net/
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=lCIv
-----END PGP SIGNATURE-----


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

A bit late in the thread but I don't see the HP E1740A -pete On 1/14/13, Jonatan Walck <jwalck@netnod.se> wrote: > -----BEGIN PGP SIGNED MESSAGE----- > Hash: SHA1 > > On 2013-01-11 02:48, Bob Camp wrote: >> HI >> >> One thing missing from the list was the HP 53181. It's useless as >> a timing device and thus didn't get listed. If you are after >> frequency only, they occasionally show up for quite a bit less than >> the 53131's. As a frequency counter, they have the same performance >> as a '131. All of the 53131, 132, 181, and 230's are small enough >> to lug around. >> >> Bob > > I'll find uses for a full TIC rather than just a fequency counter, but > I absorb all this knowledge happily either way and it's good to have > that here as a reference. > > I initially found the 53131A and everything above it to steep for my > first home lab TIC but you are right about the size, they are quite > small. (Having access to 53131's through work.) > > > As great luck would have it though, I found a second hand Pendulum > CNT-90 and I got a too good deal to hold back my trigger finger. > Picked it up yesterday! Made no measurements past "does it start up > and can it detect any frequency?" yet but will put it to good use soon. > > I wouldn't have know what to look for and find this TIC without this > list. A big thanks to all time-nuts once again! > > // jwalck > -----BEGIN PGP SIGNATURE----- > Version: GnuPG v1.4.12 (GNU/Linux) > Comment: Using GnuPG with Mozilla - http://www.enigmail.net/ > > iQIcBAEBAgAGBQJQ88oBAAoJEFwg9i9GDX+n+mMP/iWYkLQd4o3zVglTLESM8Gqu > yo9SGThQQ1Uc4svaNLzHP6WC6uO5Dh+KG6oUDhDeeqL4CNDoCCs2F4lbMsavxgW/ > lcrDjSCqATgB7J8edlLP/lw2q+Uo4VY6Y1qFmAah+6ooqU5D+mdi5j7Solm4SF+T > O+VKgghklgPuw1mWWaESNgKvv869nNkMTKw/hwQF4DDiCu/9N/gv4bxt5P778oWv > r8nu4Yck4AWWhGkDn3JcpCXqIGrA2T40CR5ek1TD9i6LGLBnCU7FmIrofd/fZEF3 > AJVYfB+tIL6CO5jj+Xvk4G4HeOXMhknKZBVmF/kFO27q4IKFAbYhIOAWdWK1MwQT > tO7N4banehXrkFuDJwJdZuze+OweME69/uBv2Rcj6jC4IWDk0kmcJKrcbGBzwuN9 > S01jyea9Ndi1CVXfdYkY9C9Pp1FZM9FzIleSB7+IGYfgS6vU+lXNnq8OVndYiT5J > 6zNjpOjd1NUWCU7WCqxLQmMEGNw1CG7ITY8G8gPAAsCsg2AkYKHix8WYidAvlc7o > NBkp7kFhD4l2jD9N8Uu50+Y2/meliiJK2wFgwbTdLOKNfdVjQYwsJ7QcQ8vJtHju > oxsESC/FVIvU0CyJ81tl7xQPV1kG9XQFuMKEISIyqCsA84+kApFP++sguM8RRLJ7 > ctm8Is0scpOVQ7kLkkmm > =lCIv > -----END PGP SIGNATURE----- > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Tue, Jan 15, 2013 5:20 PM

Hi

I'm not quite sure if the 1740 is a counter or not. I have a few of them
sitting here, but they are a bear to do anything useful with. They make the
5371 look like a portable instrument.

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Pete Lancashire
Sent: Tuesday, January 15, 2013 9:35 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Interval Timer Recommendation

A bit late in the thread but I don't see the HP E1740A

-pete

On 1/14/13, Jonatan Walck jwalck@netnod.se wrote:

-----BEGIN PGP SIGNED MESSAGE-----
Hash: SHA1

On 2013-01-11 02:48, Bob Camp wrote:

HI

One thing missing from the list was the HP 53181. It's useless as
a timing device and thus didn't get listed. If you are after
frequency only, they occasionally show up for quite a bit less than
the 53131's. As a frequency counter, they have the same performance
as a '131. All of the 53131, 132, 181, and 230's are small enough
to lug around.

Bob

I'll find uses for a full TIC rather than just a fequency counter, but
I absorb all this knowledge happily either way and it's good to have
that here as a reference.

I initially found the 53131A and everything above it to steep for my
first home lab TIC but you are right about the size, they are quite
small. (Having access to 53131's through work.)

As great luck would have it though, I found a second hand Pendulum
CNT-90 and I got a too good deal to hold back my trigger finger.
Picked it up yesterday! Made no measurements past "does it start up
and can it detect any frequency?" yet but will put it to good use soon.

I wouldn't have know what to look for and find this TIC without this
list. A big thanks to all time-nuts once again!

// jwalck
-----BEGIN PGP SIGNATURE-----
Version: GnuPG v1.4.12 (GNU/Linux)
Comment: Using GnuPG with Mozilla - http://www.enigmail.net/
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=lCIv
-----END PGP SIGNATURE-----


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I'm not quite sure if the 1740 is a counter or not. I have a few of them sitting here, but they are a bear to do anything useful with. They make the 5371 look like a portable instrument. Bob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Pete Lancashire Sent: Tuesday, January 15, 2013 9:35 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] Interval Timer Recommendation A bit late in the thread but I don't see the HP E1740A -pete On 1/14/13, Jonatan Walck <jwalck@netnod.se> wrote: > -----BEGIN PGP SIGNED MESSAGE----- > Hash: SHA1 > > On 2013-01-11 02:48, Bob Camp wrote: >> HI >> >> One thing missing from the list was the HP 53181. It's useless as >> a timing device and thus didn't get listed. If you are after >> frequency only, they occasionally show up for quite a bit less than >> the 53131's. As a frequency counter, they have the same performance >> as a '131. All of the 53131, 132, 181, and 230's are small enough >> to lug around. >> >> Bob > > I'll find uses for a full TIC rather than just a fequency counter, but > I absorb all this knowledge happily either way and it's good to have > that here as a reference. > > I initially found the 53131A and everything above it to steep for my > first home lab TIC but you are right about the size, they are quite > small. (Having access to 53131's through work.) > > > As great luck would have it though, I found a second hand Pendulum > CNT-90 and I got a too good deal to hold back my trigger finger. > Picked it up yesterday! Made no measurements past "does it start up > and can it detect any frequency?" yet but will put it to good use soon. > > I wouldn't have know what to look for and find this TIC without this > list. A big thanks to all time-nuts once again! > > // jwalck > -----BEGIN PGP SIGNATURE----- > Version: GnuPG v1.4.12 (GNU/Linux) > Comment: Using GnuPG with Mozilla - http://www.enigmail.net/ > > iQIcBAEBAgAGBQJQ88oBAAoJEFwg9i9GDX+n+mMP/iWYkLQd4o3zVglTLESM8Gqu > yo9SGThQQ1Uc4svaNLzHP6WC6uO5Dh+KG6oUDhDeeqL4CNDoCCs2F4lbMsavxgW/ > lcrDjSCqATgB7J8edlLP/lw2q+Uo4VY6Y1qFmAah+6ooqU5D+mdi5j7Solm4SF+T > O+VKgghklgPuw1mWWaESNgKvv869nNkMTKw/hwQF4DDiCu/9N/gv4bxt5P778oWv > r8nu4Yck4AWWhGkDn3JcpCXqIGrA2T40CR5ek1TD9i6LGLBnCU7FmIrofd/fZEF3 > AJVYfB+tIL6CO5jj+Xvk4G4HeOXMhknKZBVmF/kFO27q4IKFAbYhIOAWdWK1MwQT > tO7N4banehXrkFuDJwJdZuze+OweME69/uBv2Rcj6jC4IWDk0kmcJKrcbGBzwuN9 > S01jyea9Ndi1CVXfdYkY9C9Pp1FZM9FzIleSB7+IGYfgS6vU+lXNnq8OVndYiT5J > 6zNjpOjd1NUWCU7WCqxLQmMEGNw1CG7ITY8G8gPAAsCsg2AkYKHix8WYidAvlc7o > NBkp7kFhD4l2jD9N8Uu50+Y2/meliiJK2wFgwbTdLOKNfdVjQYwsJ7QcQ8vJtHju > oxsESC/FVIvU0CyJ81tl7xQPV1kG9XQFuMKEISIyqCsA84+kApFP++sguM8RRLJ7 > ctm8Is0scpOVQ7kLkkmm > =lCIv > -----END PGP SIGNATURE----- > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.