time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Zeeman frequency oddness

JA
John Ackermann N8UR
Sat, Oct 27, 2012 7:16 PM

Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd
go through the whole setup routine for my 5061B/004 and see how close
the C-field-via-Zeeman setting would bring me to GPS-derived frequency.
It turned out to be an interesting and puzzling exercise.

This 5061 seems to be in perfect working order -- quick lock, good meter
readings, and measurements indicate frequency within parts in e12.  I
don't have any reason to believe that it's not tuned or working
properly, except for the Zeeman-setting results.

The problem is that when I tune the audio source around 53.53 kHz, per
both the manual and the sticker on the door, I don't see any change in
Beam I at all.  Nor do I see anything at the alternate frequency of
42.82 kHz.

Instead, I see the expected three peaks -- primary with a smaller
secondary on either side -- at about 48.21 kHz, which doesn't show up
anywhere in the literature I've found.

Below, I've cut and pasted a years-old message from TVB and Corby that
explains the Zeeman frequencies.  I've measured the synthesizer output
and it's nominally 12.6317725 MHz, which per that message should
correspond to a 53.53 Zeeman.  Where 48.21 kHz comes from, I have no idea.

I'm using a Rigol arbitrary function generator locked to an external
reference as the audio source, in sine wave mode.  I know that's not the
cleanest device in the world, but the wave doesn't look too bad on my
scope and a counter indicates the frequency is what the dial says.  As I
adjust the audio amplitude, the beam current responds, and I see a peak
at around 500mV, which the 5061B manual says is correct.

Any ideas why I might be seeing this very off-the-wall result?  Could
distortion in the audio source cause something like this?  I'm more
inclined to blame technique or gremlins than the 5061B -- again,
external measurements indicate that the thing is tuned correctly and
operating properly, just having this goofy Zeeman response.

Thanks!

John

[time-nuts] Zeeman frequency and cesium tube interchange
Tom Van Baak
Fri Apr 22 13:32:35 EDT 2005

Hi Brian,

The SI second is defined for mean sea level and no
external fields.

If there were no magnetic field, in theory, a cesium
tube would show a resonance when its synthesizer
generated exactly 9192.631770 MHz; the definition
of the second.

But in practice, a weak uniform magnetic field is
necessary for the beam apparatus to operate; to
isolate the center peak from the other peaks. This
DC field also has the side-effect of slightly shifting
the frequency of the center line.

Fortunately the frequency shift is a calculatable
amount (a function of magnetic field strength) so
the trick is that the synthesizer must be designed
to generate a slightly higher frequency to exactly
compensate for the shift that will be induced by
the field.

Thus the synthesizer for a 5061A does not actually
generate 9192 631 770 Hz as one might expect,
but because of the nominal 61 milligauss C-field,
the synthesizer must generate 9192 631 771.6 Hz
in order to lock onto the Cs peak precisely. The
Zeeman frequency for 61 mG is 42.82 kHz (info
from an old 5061A manual).

Other 5061A/B use a 76 mG field, corresponding
to a 53.53 kHz Zeeman frequency, and require the
synthesizer to generate 9192 631 772.5 Hz (info
from a new 5061B manual).

The short HP 5062C runs at 9192 631 774.3 Hz
with a Zeeman of 70.40 kHz.

Below is a great reply from Corby Dawson about
the problems this can cause when mixing FTS
4050, 4060, HP 5060A, 5061A, 5061B parts.

/tvb

----- Original Message -----

Tom,

The zeeman frequency required depends on two things, the magnitude of the
C-field current and the synthesizer frequency.

HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are
configured for a lower value of C-field current by selecting a higher
value series resistor on the A15 board. In this case any tube installed
(5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency.

HP 5061A and B units and 5060A units that have a synthesizer frequency of
12.6317716 Mhz are configured for a higher C-field current due to a lower
value series resistor in the C-field circuit. In this case any installed
tube will operate with a 42.82Khz zeeman frequency.

Problems arise when synthesizer and or A15 modules are swapped around
indiscriminantly leaving a unit with modules that do not match!

Since the HP and FTS tubes C-field windings are designed to provide the
same field for a given input they are interchangeable. You can operate
any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick
with how the mainframe came configured.

I don't remember what I ended up with as far as the zeeman freq. was
concerned when I installed a Frequency Electronics tube into a 5061A, but
do remember the line width was quite broad in keeping with the reduced
accuracy spec. of the FE tubes. (Same spec. as the 5062C tube)

I have installed FTS tubes into the 5062C giving it the +-7Z10-12th
accuracy spec. and did have to modify the synthesizer for 42.82Khz and
set the C-field accordingly.

Hope this helps!

Corby_

Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd go through the whole setup routine for my 5061B/004 and see how close the C-field-via-Zeeman setting would bring me to GPS-derived frequency. It turned out to be an interesting and puzzling exercise. This 5061 seems to be in perfect working order -- quick lock, good meter readings, and measurements indicate frequency within parts in e12. I don't have any reason to believe that it's not tuned or working properly, except for the Zeeman-setting results. The problem is that when I tune the audio source around 53.53 kHz, per both the manual and the sticker on the door, I don't see any change in Beam I at all. Nor do I see anything at the alternate frequency of 42.82 kHz. Instead, I see the expected three peaks -- primary with a smaller secondary on either side -- at about 48.21 kHz, which doesn't show up anywhere in the literature I've found. Below, I've cut and pasted a years-old message from TVB and Corby that explains the Zeeman frequencies. I've measured the synthesizer output and it's nominally 12.6317725 MHz, which per that message should correspond to a 53.53 Zeeman. Where 48.21 kHz comes from, I have no idea. I'm using a Rigol arbitrary function generator locked to an external reference as the audio source, in sine wave mode. I know that's not the cleanest device in the world, but the wave doesn't look too bad on my scope and a counter indicates the frequency is what the dial says. As I adjust the audio amplitude, the beam current responds, and I see a peak at around 500mV, which the 5061B manual says is correct. Any ideas why I might be seeing this very off-the-wall result? Could distortion in the audio source cause something like this? I'm more inclined to blame technique or gremlins than the 5061B -- again, external measurements indicate that the thing is tuned correctly and operating properly, just having this goofy Zeeman response. Thanks! John [time-nuts] Zeeman frequency and cesium tube interchange Tom Van Baak Fri Apr 22 13:32:35 EDT 2005 Hi Brian, The SI second is defined for mean sea level and no external fields. If there were no magnetic field, in theory, a cesium tube would show a resonance when its synthesizer generated exactly 9192.631770 MHz; the definition of the second. But in practice, a weak uniform magnetic field is necessary for the beam apparatus to operate; to isolate the center peak from the other peaks. This DC field also has the side-effect of slightly shifting the frequency of the center line. Fortunately the frequency shift is a calculatable amount (a function of magnetic field strength) so the trick is that the synthesizer must be designed to generate a slightly higher frequency to exactly compensate for the shift that will be induced by the field. Thus the synthesizer for a 5061A does not actually generate 9192 631 770 Hz as one might expect, but because of the nominal 61 milligauss C-field, the synthesizer must generate 9192 631 771.6 Hz in order to lock onto the Cs peak precisely. The Zeeman frequency for 61 mG is 42.82 kHz (info from an old 5061A manual). Other 5061A/B use a 76 mG field, corresponding to a 53.53 kHz Zeeman frequency, and require the synthesizer to generate 9192 631 772.5 Hz (info from a new 5061B manual). The short HP 5062C runs at 9192 631 774.3 Hz with a Zeeman of 70.40 kHz. Below is a great reply from Corby Dawson about the problems this can cause when mixing FTS 4050, 4060, HP 5060A, 5061A, 5061B parts. /tvb ----- Original Message ----- > Tom, > > The zeeman frequency required depends on two things, the magnitude of the > C-field current and the synthesizer frequency. > > HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are > configured for a lower value of C-field current by selecting a higher > value series resistor on the A15 board. In this case any tube installed > (5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency. > > HP 5061A and B units and 5060A units that have a synthesizer frequency of > 12.6317716 Mhz are configured for a higher C-field current due to a lower > value series resistor in the C-field circuit. In this case any installed > tube will operate with a 42.82Khz zeeman frequency. > > Problems arise when synthesizer and or A15 modules are swapped around > indiscriminantly leaving a unit with modules that do not match! > > Since the HP and FTS tubes C-field windings are designed to provide the > same field for a given input they are interchangeable. You can operate > any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick > with how the mainframe came configured. > > > I don't remember what I ended up with as far as the zeeman freq. was > concerned when I installed a Frequency Electronics tube into a 5061A, but > do remember the line width was quite broad in keeping with the reduced > accuracy spec. of the FE tubes. (Same spec. as the 5062C tube) > > I have installed FTS tubes into the 5062C giving it the +-7Z10-12th > accuracy spec. and did have to modify the synthesizer for 42.82Khz and > set the C-field accordingly. > > Hope this helps! > > Corby_
PS
paul swed
Sat, Oct 27, 2012 8:55 PM

John
Going from what you have here and the information below it would seem to me
that there is some form of residual magnetism effecting the tube. I don't
believe the waveform of the signal has a huge effect as long as its not a
square wave. I think it can have distortion.

Its been a while but since you have an accurate reference, I believe you
can adjust the system to align to that and see what pip on the zeeman you
end up on. Perhaps you are aligning to the wrong pip. Go to either of the
lower ones and see if it works better or lines up. If it does then the math
holds and refines the question.

There was an old HP doc that evidently it was very easy to align to the
wrong pip. It was quite a problem actually.
Regards
Paul
WB8TSL

On Sat, Oct 27, 2012 at 3:16 PM, John Ackermann N8UR jra@febo.com wrote:

Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd go
through the whole setup routine for my 5061B/004 and see how close the
C-field-via-Zeeman setting would bring me to GPS-derived frequency.  It
turned out to be an interesting and puzzling exercise.

This 5061 seems to be in perfect working order -- quick lock, good meter
readings, and measurements indicate frequency within parts in e12.  I don't
have any reason to believe that it's not tuned or working properly, except
for the Zeeman-setting results.

The problem is that when I tune the audio source around 53.53 kHz, per
both the manual and the sticker on the door, I don't see any change in Beam
I at all.  Nor do I see anything at the alternate frequency of 42.82 kHz.

Instead, I see the expected three peaks -- primary with a smaller
secondary on either side -- at about 48.21 kHz, which doesn't show up
anywhere in the literature I've found.

Below, I've cut and pasted a years-old message from TVB and Corby that
explains the Zeeman frequencies.  I've measured the synthesizer output and
it's nominally 12.6317725 MHz, which per that message should correspond to
a 53.53 Zeeman.  Where 48.21 kHz comes from, I have no idea.

I'm using a Rigol arbitrary function generator locked to an external
reference as the audio source, in sine wave mode.  I know that's not the
cleanest device in the world, but the wave doesn't look too bad on my scope
and a counter indicates the frequency is what the dial says.  As I adjust
the audio amplitude, the beam current responds, and I see a peak at around
500mV, which the 5061B manual says is correct.

Any ideas why I might be seeing this very off-the-wall result?  Could
distortion in the audio source cause something like this?  I'm more
inclined to blame technique or gremlins than the 5061B -- again, external
measurements indicate that the thing is tuned correctly and operating
properly, just having this goofy Zeeman response.

Thanks!

John

[time-nuts] Zeeman frequency and cesium tube interchange
Tom Van Baak
Fri Apr 22 13:32:35 EDT 2005

Hi Brian,

The SI second is defined for mean sea level and no
external fields.

If there were no magnetic field, in theory, a cesium
tube would show a resonance when its synthesizer
generated exactly 9192.631770 MHz; the definition
of the second.

But in practice, a weak uniform magnetic field is
necessary for the beam apparatus to operate; to
isolate the center peak from the other peaks. This
DC field also has the side-effect of slightly shifting
the frequency of the center line.

Fortunately the frequency shift is a calculatable
amount (a function of magnetic field strength) so
the trick is that the synthesizer must be designed
to generate a slightly higher frequency to exactly
compensate for the shift that will be induced by
the field.

Thus the synthesizer for a 5061A does not actually
generate 9192 631 770 Hz as one might expect,
but because of the nominal 61 milligauss C-field,
the synthesizer must generate 9192 631 771.6 Hz
in order to lock onto the Cs peak precisely. The
Zeeman frequency for 61 mG is 42.82 kHz (info
from an old 5061A manual).

Other 5061A/B use a 76 mG field, corresponding
to a 53.53 kHz Zeeman frequency, and require the
synthesizer to generate 9192 631 772.5 Hz (info
from a new 5061B manual).

The short HP 5062C runs at 9192 631 774.3 Hz
with a Zeeman of 70.40 kHz.

Below is a great reply from Corby Dawson about
the problems this can cause when mixing FTS
4050, 4060, HP 5060A, 5061A, 5061B parts.

/tvb

----- Original Message -----

Tom,

The zeeman frequency required depends on two things, the magnitude of the
C-field current and the synthesizer frequency.

HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are
configured for a lower value of C-field current by selecting a higher
value series resistor on the A15 board. In this case any tube installed
(5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency.

HP 5061A and B units and 5060A units that have a synthesizer frequency of
12.6317716 Mhz are configured for a higher C-field current due to a lower
value series resistor in the C-field circuit. In this case any installed
tube will operate with a 42.82Khz zeeman frequency.

Problems arise when synthesizer and or A15 modules are swapped around
indiscriminantly leaving a unit with modules that do not match!

Since the HP and FTS tubes C-field windings are designed to provide the
same field for a given input they are interchangeable. You can operate
any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick
with how the mainframe came configured.

I don't remember what I ended up with as far as the zeeman freq. was
concerned when I installed a Frequency Electronics tube into a 5061A, but
do remember the line width was quite broad in keeping with the reduced
accuracy spec. of the FE tubes. (Same spec. as the 5062C tube)

I have installed FTS tubes into the 5062C giving it the +-7Z10-12th
accuracy spec. and did have to modify the synthesizer for 42.82Khz and
set the C-field accordingly.

Hope this helps!

Corby_

_____________**
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**
mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

John Going from what you have here and the information below it would seem to me that there is some form of residual magnetism effecting the tube. I don't believe the waveform of the signal has a huge effect as long as its not a square wave. I think it can have distortion. Its been a while but since you have an accurate reference, I believe you can adjust the system to align to that and see what pip on the zeeman you end up on. Perhaps you are aligning to the wrong pip. Go to either of the lower ones and see if it works better or lines up. If it does then the math holds and refines the question. There was an old HP doc that evidently it was very easy to align to the wrong pip. It was quite a problem actually. Regards Paul WB8TSL On Sat, Oct 27, 2012 at 3:16 PM, John Ackermann N8UR <jra@febo.com> wrote: > Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd go > through the whole setup routine for my 5061B/004 and see how close the > C-field-via-Zeeman setting would bring me to GPS-derived frequency. It > turned out to be an interesting and puzzling exercise. > > This 5061 seems to be in perfect working order -- quick lock, good meter > readings, and measurements indicate frequency within parts in e12. I don't > have any reason to believe that it's not tuned or working properly, except > for the Zeeman-setting results. > > The problem is that when I tune the audio source around 53.53 kHz, per > both the manual and the sticker on the door, I don't see any change in Beam > I at all. Nor do I see anything at the alternate frequency of 42.82 kHz. > > Instead, I see the expected three peaks -- primary with a smaller > secondary on either side -- at about 48.21 kHz, which doesn't show up > anywhere in the literature I've found. > > Below, I've cut and pasted a years-old message from TVB and Corby that > explains the Zeeman frequencies. I've measured the synthesizer output and > it's nominally 12.6317725 MHz, which per that message should correspond to > a 53.53 Zeeman. Where 48.21 kHz comes from, I have no idea. > > I'm using a Rigol arbitrary function generator locked to an external > reference as the audio source, in sine wave mode. I know that's not the > cleanest device in the world, but the wave doesn't look too bad on my scope > and a counter indicates the frequency is what the dial says. As I adjust > the audio amplitude, the beam current responds, and I see a peak at around > 500mV, which the 5061B manual says is correct. > > Any ideas why I might be seeing this very off-the-wall result? Could > distortion in the audio source cause something like this? I'm more > inclined to blame technique or gremlins than the 5061B -- again, external > measurements indicate that the thing is tuned correctly and operating > properly, just having this goofy Zeeman response. > > Thanks! > > John > > > [time-nuts] Zeeman frequency and cesium tube interchange > Tom Van Baak > Fri Apr 22 13:32:35 EDT 2005 > > Hi Brian, > > The SI second is defined for mean sea level and no > external fields. > > If there were no magnetic field, in theory, a cesium > tube would show a resonance when its synthesizer > generated exactly 9192.631770 MHz; the definition > of the second. > > But in practice, a weak uniform magnetic field is > necessary for the beam apparatus to operate; to > isolate the center peak from the other peaks. This > DC field also has the side-effect of slightly shifting > the frequency of the center line. > > Fortunately the frequency shift is a calculatable > amount (a function of magnetic field strength) so > the trick is that the synthesizer must be designed > to generate a slightly higher frequency to exactly > compensate for the shift that will be induced by > the field. > > Thus the synthesizer for a 5061A does not actually > generate 9192 631 770 Hz as one might expect, > but because of the nominal 61 milligauss C-field, > the synthesizer must generate 9192 631 771.6 Hz > in order to lock onto the Cs peak precisely. The > Zeeman frequency for 61 mG is 42.82 kHz (info > from an old 5061A manual). > > Other 5061A/B use a 76 mG field, corresponding > to a 53.53 kHz Zeeman frequency, and require the > synthesizer to generate 9192 631 772.5 Hz (info > from a new 5061B manual). > > The short HP 5062C runs at 9192 631 774.3 Hz > with a Zeeman of 70.40 kHz. > > Below is a great reply from Corby Dawson about > the problems this can cause when mixing FTS > 4050, 4060, HP 5060A, 5061A, 5061B parts. > > /tvb > > ----- Original Message ----- > > Tom, > > > > The zeeman frequency required depends on two things, the magnitude of the > > C-field current and the synthesizer frequency. > > > > HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are > > configured for a lower value of C-field current by selecting a higher > > value series resistor on the A15 board. In this case any tube installed > > (5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency. > > > > HP 5061A and B units and 5060A units that have a synthesizer frequency of > > 12.6317716 Mhz are configured for a higher C-field current due to a lower > > value series resistor in the C-field circuit. In this case any installed > > tube will operate with a 42.82Khz zeeman frequency. > > > > Problems arise when synthesizer and or A15 modules are swapped around > > indiscriminantly leaving a unit with modules that do not match! > > > > Since the HP and FTS tubes C-field windings are designed to provide the > > same field for a given input they are interchangeable. You can operate > > any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick > > with how the mainframe came configured. > > > > > > I don't remember what I ended up with as far as the zeeman freq. was > > concerned when I installed a Frequency Electronics tube into a 5061A, but > > do remember the line width was quite broad in keeping with the reduced > > accuracy spec. of the FE tubes. (Same spec. as the 5062C tube) > > > > I have installed FTS tubes into the 5062C giving it the +-7Z10-12th > > accuracy spec. and did have to modify the synthesizer for 42.82Khz and > > set the C-field accordingly. > > > > Hope this helps! > > > > Corby_ > > ______________________________**_________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/time-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. >
MD
Magnus Danielson
Sat, Oct 27, 2012 10:30 PM

Hi John,

Don't know if I address your real issues, but I want to add some more
pieces to the puzzle.

On 10/27/2012 09:16 PM, John Ackermann N8UR wrote:

Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd
go through the whole setup routine for my 5061B/004 and see how close
the C-field-via-Zeeman setting would bring me to GPS-derived frequency.
It turned out to be an interesting and puzzling exercise.

This 5061 seems to be in perfect working order -- quick lock, good meter
readings, and measurements indicate frequency within parts in e12. I
don't have any reason to believe that it's not tuned or working
properly, except for the Zeeman-setting results.

The problem is that when I tune the audio source around 53.53 kHz, per
both the manual and the sticker on the door, I don't see any change in
Beam I at all. Nor do I see anything at the alternate frequency of 42.82
kHz.

Instead, I see the expected three peaks -- primary with a smaller
secondary on either side -- at about 48.21 kHz, which doesn't show up
anywhere in the literature I've found.

There are in total 7 peaks, you want the center peak of those.

Tom has made measurements:
http://leapsecond.com/images/cfield.gif

I have done the same to one of my tubes, but I don't recall where I have
that flimsy picture, so the above is a good start.

Below, I've cut and pasted a years-old message from TVB and Corby that
explains the Zeeman frequencies. I've measured the synthesizer output
and it's nominally 12.6317725 MHz, which per that message should
correspond to a 53.53 Zeeman. Where 48.21 kHz comes from, I have no idea.

The separation of the peaks depends on the C-field value you have.

For low C-field strength, the side-peaks separate my the square of the
C-field (B), and the cluster shifts gently linear with the C-field.

You want to spread the side-peaks out, such that they do not confuse
your measures. Also, looking at the above you have the 7 Rabi
distributions, and on top of those the Ramsay fringes, at which you want
to lock onto the center one. It may be interesting to learn that certain
systematics skews the shape of these, and thus causes a systematic
miss-tuning, so great care is taken to reduce that effect when
manufacturing the tube.

Look at the above C-field plot again and you see how the side features
move with different C-field settings.

Modern "digital" caesium clocks measure the side-features in order to
Servo the C-field into a stable value, and hence also stabilize C-field
drift out of the equation first degree.

I'm using a Rigol arbitrary function generator locked to an external
reference as the audio source, in sine wave mode. I know that's not the
cleanest device in the world, but the wave doesn't look too bad on my
scope and a counter indicates the frequency is what the dial says. As I
adjust the audio amplitude, the beam current responds, and I see a peak
at around 500mV, which the 5061B manual says is correct.

Any ideas why I might be seeing this very off-the-wall result? Could
distortion in the audio source cause something like this? I'm more
inclined to blame technique or gremlins than the 5061B -- again,
external measurements indicate that the thing is tuned correctly and
operating properly, just having this goofy Zeeman response.

You want a clean source, as spurs or distortion would cause you to look
at multiple points in the spectrum at the same time and you would get
the combined result of those features. Still, the distortion values
doesn't have to be stellar to get decent readings.

Cheers,
Magnus

Hi John, Don't know if I address your real issues, but I want to add some more pieces to the puzzle. On 10/27/2012 09:16 PM, John Ackermann N8UR wrote: > Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd > go through the whole setup routine for my 5061B/004 and see how close > the C-field-via-Zeeman setting would bring me to GPS-derived frequency. > It turned out to be an interesting and puzzling exercise. > > This 5061 seems to be in perfect working order -- quick lock, good meter > readings, and measurements indicate frequency within parts in e12. I > don't have any reason to believe that it's not tuned or working > properly, except for the Zeeman-setting results. > > The problem is that when I tune the audio source around 53.53 kHz, per > both the manual and the sticker on the door, I don't see any change in > Beam I at all. Nor do I see anything at the alternate frequency of 42.82 > kHz. > > Instead, I see the expected three peaks -- primary with a smaller > secondary on either side -- at about 48.21 kHz, which doesn't show up > anywhere in the literature I've found. There are in total 7 peaks, you want the center peak of those. Tom has made measurements: http://leapsecond.com/images/cfield.gif I have done the same to one of my tubes, but I don't recall where I have that flimsy picture, so the above is a good start. > Below, I've cut and pasted a years-old message from TVB and Corby that > explains the Zeeman frequencies. I've measured the synthesizer output > and it's nominally 12.6317725 MHz, which per that message should > correspond to a 53.53 Zeeman. Where 48.21 kHz comes from, I have no idea. The separation of the peaks depends on the C-field value you have. For low C-field strength, the side-peaks separate my the square of the C-field (B), and the cluster shifts gently linear with the C-field. You want to spread the side-peaks out, such that they do not confuse your measures. Also, looking at the above you have the 7 Rabi distributions, and on top of those the Ramsay fringes, at which you want to lock onto the center one. It may be interesting to learn that certain systematics skews the shape of these, and thus causes a systematic miss-tuning, so great care is taken to reduce that effect when manufacturing the tube. Look at the above C-field plot again and you see how the side features move with different C-field settings. Modern "digital" caesium clocks measure the side-features in order to Servo the C-field into a stable value, and hence also stabilize C-field drift out of the equation first degree. > I'm using a Rigol arbitrary function generator locked to an external > reference as the audio source, in sine wave mode. I know that's not the > cleanest device in the world, but the wave doesn't look too bad on my > scope and a counter indicates the frequency is what the dial says. As I > adjust the audio amplitude, the beam current responds, and I see a peak > at around 500mV, which the 5061B manual says is correct. > > Any ideas why I might be seeing this very off-the-wall result? Could > distortion in the audio source cause something like this? I'm more > inclined to blame technique or gremlins than the 5061B -- again, > external measurements indicate that the thing is tuned correctly and > operating properly, just having this goofy Zeeman response. You want a clean source, as spurs or distortion would cause you to look at multiple points in the spectrum at the same time and you would get the combined result of those features. Still, the distortion values doesn't have to be stellar to get decent readings. Cheers, Magnus
CH
Chuck Harris
Sat, Oct 27, 2012 11:13 PM

Two things come to mind:

  1. is this the high performance tube?  Perhaps it has a different
    Zeeman frequency than the standard tube?
  2. a non HP replacement tube?  FTS Cs beam tubes are way different
    Zeeman frequencies than HP.

-Chuck Harris

John Ackermann N8UR wrote:

Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd go through the
whole setup routine for my 5061B/004 and see how close the C-field-via-Zeeman setting
would bring me to GPS-derived frequency.  It turned out to be an interesting and
puzzling exercise.

This 5061 seems to be in perfect working order -- quick lock, good meter readings,
and measurements indicate frequency within parts in e12.  I don't have any reason to
believe that it's not tuned or working properly, except for the Zeeman-setting results.

The problem is that when I tune the audio source around 53.53 kHz, per both the
manual and the sticker on the door, I don't see any change in Beam I at all.  Nor do
I see anything at the alternate frequency of 42.82 kHz.

Instead, I see the expected three peaks -- primary with a smaller secondary on either
side -- at about 48.21 kHz, which doesn't show up anywhere in the literature I've found.

Below, I've cut and pasted a years-old message from TVB and Corby that explains the
Zeeman frequencies.  I've measured the synthesizer output and it's nominally
12.6317725 MHz, which per that message should correspond to a 53.53 Zeeman.  Where
48.21 kHz comes from, I have no idea.

I'm using a Rigol arbitrary function generator locked to an external reference as the
audio source, in sine wave mode.  I know that's not the cleanest device in the world,
but the wave doesn't look too bad on my scope and a counter indicates the frequency
is what the dial says.  As I adjust the audio amplitude, the beam current responds,
and I see a peak at around 500mV, which the 5061B manual says is correct.

Any ideas why I might be seeing this very off-the-wall result?  Could distortion in
the audio source cause something like this?  I'm more inclined to blame technique or
gremlins than the 5061B -- again, external measurements indicate that the thing is
tuned correctly and operating properly, just having this goofy Zeeman response.

Thanks!

John

[time-nuts] Zeeman frequency and cesium tube interchange
Tom Van Baak
Fri Apr 22 13:32:35 EDT 2005

Hi Brian,

The SI second is defined for mean sea level and no
external fields.

If there were no magnetic field, in theory, a cesium
tube would show a resonance when its synthesizer
generated exactly 9192.631770 MHz; the definition
of the second.

But in practice, a weak uniform magnetic field is
necessary for the beam apparatus to operate; to
isolate the center peak from the other peaks. This
DC field also has the side-effect of slightly shifting
the frequency of the center line.

Fortunately the frequency shift is a calculatable
amount (a function of magnetic field strength) so
the trick is that the synthesizer must be designed
to generate a slightly higher frequency to exactly
compensate for the shift that will be induced by
the field.

Thus the synthesizer for a 5061A does not actually
generate 9192 631 770 Hz as one might expect,
but because of the nominal 61 milligauss C-field,
the synthesizer must generate 9192 631 771.6 Hz
in order to lock onto the Cs peak precisely. The
Zeeman frequency for 61 mG is 42.82 kHz (info
from an old 5061A manual).

Other 5061A/B use a 76 mG field, corresponding
to a 53.53 kHz Zeeman frequency, and require the
synthesizer to generate 9192 631 772.5 Hz (info
from a new 5061B manual).

The short HP 5062C runs at 9192 631 774.3 Hz
with a Zeeman of 70.40 kHz.

Below is a great reply from Corby Dawson about
the problems this can cause when mixing FTS
4050, 4060, HP 5060A, 5061A, 5061B parts.

/tvb

----- Original Message -----

Tom,

The zeeman frequency required depends on two things, the magnitude of the
C-field current and the synthesizer frequency.

HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are
configured for a lower value of C-field current by selecting a higher
value series resistor on the A15 board. In this case any tube installed
(5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency.

HP 5061A and B units and 5060A units that have a synthesizer frequency of
12.6317716 Mhz are configured for a higher C-field current due to a lower
value series resistor in the C-field circuit. In this case any installed
tube will operate with a 42.82Khz zeeman frequency.

Problems arise when synthesizer and or A15 modules are swapped around
indiscriminantly leaving a unit with modules that do not match!

Since the HP and FTS tubes C-field windings are designed to provide the
same field for a given input they are interchangeable. You can operate
any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick
with how the mainframe came configured.

I don't remember what I ended up with as far as the zeeman freq. was
concerned when I installed a Frequency Electronics tube into a 5061A, but
do remember the line width was quite broad in keeping with the reduced
accuracy spec. of the FE tubes. (Same spec. as the 5062C tube)

I have installed FTS tubes into the 5062C giving it the +-7Z10-12th
accuracy spec. and did have to modify the synthesizer for 42.82Khz and
set the C-field accordingly.

Hope this helps!

Corby_


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Two things come to mind: 1) is this the high performance tube? Perhaps it has a different Zeeman frequency than the standard tube? 2) a non HP replacement tube? FTS Cs beam tubes are way different Zeeman frequencies than HP. -Chuck Harris John Ackermann N8UR wrote: > Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd go through the > whole setup routine for my 5061B/004 and see how close the C-field-via-Zeeman setting > would bring me to GPS-derived frequency. It turned out to be an interesting and > puzzling exercise. > > This 5061 seems to be in perfect working order -- quick lock, good meter readings, > and measurements indicate frequency within parts in e12. I don't have any reason to > believe that it's not tuned or working properly, except for the Zeeman-setting results. > > The problem is that when I tune the audio source around 53.53 kHz, per both the > manual and the sticker on the door, I don't see any change in Beam I at all. Nor do > I see anything at the alternate frequency of 42.82 kHz. > > Instead, I see the expected three peaks -- primary with a smaller secondary on either > side -- at about 48.21 kHz, which doesn't show up anywhere in the literature I've found. > > Below, I've cut and pasted a years-old message from TVB and Corby that explains the > Zeeman frequencies. I've measured the synthesizer output and it's nominally > 12.6317725 MHz, which per that message should correspond to a 53.53 Zeeman. Where > 48.21 kHz comes from, I have no idea. > > I'm using a Rigol arbitrary function generator locked to an external reference as the > audio source, in sine wave mode. I know that's not the cleanest device in the world, > but the wave doesn't look too bad on my scope and a counter indicates the frequency > is what the dial says. As I adjust the audio amplitude, the beam current responds, > and I see a peak at around 500mV, which the 5061B manual says is correct. > > Any ideas why I might be seeing this very off-the-wall result? Could distortion in > the audio source cause something like this? I'm more inclined to blame technique or > gremlins than the 5061B -- again, external measurements indicate that the thing is > tuned correctly and operating properly, just having this goofy Zeeman response. > > Thanks! > > John > > > [time-nuts] Zeeman frequency and cesium tube interchange > Tom Van Baak > Fri Apr 22 13:32:35 EDT 2005 > > Hi Brian, > > The SI second is defined for mean sea level and no > external fields. > > If there were no magnetic field, in theory, a cesium > tube would show a resonance when its synthesizer > generated exactly 9192.631770 MHz; the definition > of the second. > > But in practice, a weak uniform magnetic field is > necessary for the beam apparatus to operate; to > isolate the center peak from the other peaks. This > DC field also has the side-effect of slightly shifting > the frequency of the center line. > > Fortunately the frequency shift is a calculatable > amount (a function of magnetic field strength) so > the trick is that the synthesizer must be designed > to generate a slightly higher frequency to exactly > compensate for the shift that will be induced by > the field. > > Thus the synthesizer for a 5061A does not actually > generate 9192 631 770 Hz as one might expect, > but because of the nominal 61 milligauss C-field, > the synthesizer must generate 9192 631 771.6 Hz > in order to lock onto the Cs peak precisely. The > Zeeman frequency for 61 mG is 42.82 kHz (info > from an old 5061A manual). > > Other 5061A/B use a 76 mG field, corresponding > to a 53.53 kHz Zeeman frequency, and require the > synthesizer to generate 9192 631 772.5 Hz (info > from a new 5061B manual). > > The short HP 5062C runs at 9192 631 774.3 Hz > with a Zeeman of 70.40 kHz. > > Below is a great reply from Corby Dawson about > the problems this can cause when mixing FTS > 4050, 4060, HP 5060A, 5061A, 5061B parts. > > /tvb > > ----- Original Message ----- > > Tom, > > > > The zeeman frequency required depends on two things, the magnitude of the > > C-field current and the synthesizer frequency. > > > > HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are > > configured for a lower value of C-field current by selecting a higher > > value series resistor on the A15 board. In this case any tube installed > > (5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency. > > > > HP 5061A and B units and 5060A units that have a synthesizer frequency of > > 12.6317716 Mhz are configured for a higher C-field current due to a lower > > value series resistor in the C-field circuit. In this case any installed > > tube will operate with a 42.82Khz zeeman frequency. > > > > Problems arise when synthesizer and or A15 modules are swapped around > > indiscriminantly leaving a unit with modules that do not match! > > > > Since the HP and FTS tubes C-field windings are designed to provide the > > same field for a given input they are interchangeable. You can operate > > any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick > > with how the mainframe came configured. > > > > > > I don't remember what I ended up with as far as the zeeman freq. was > > concerned when I installed a Frequency Electronics tube into a 5061A, but > > do remember the line width was quite broad in keeping with the reduced > > accuracy spec. of the FE tubes. (Same spec. as the 5062C tube) > > > > I have installed FTS tubes into the 5062C giving it the +-7Z10-12th > > accuracy spec. and did have to modify the synthesizer for 42.82Khz and > > set the C-field accordingly. > > > > Hope this helps! > > > > Corby_ > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
JL
J. L. Trantham
Sat, Oct 27, 2012 11:28 PM

John,

Did you do the Zeeman frequency adjustment before doing the 'degauss'?  In
other words, did you 'magnetize' rather than 'de-magnetize'?

Have you checked the A15 board to make sure it has the correct resistor and
current?  I have a 5061B Service Manual if you need some sections or
schematics.

Good luck.

Joe

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of John Ackermann N8UR
Sent: Saturday, October 27, 2012 2:16 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Zeeman frequency oddness

Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd
go through the whole setup routine for my 5061B/004 and see how close
the C-field-via-Zeeman setting would bring me to GPS-derived frequency.
It turned out to be an interesting and puzzling exercise.

This 5061 seems to be in perfect working order -- quick lock, good meter
readings, and measurements indicate frequency within parts in e12.  I
don't have any reason to believe that it's not tuned or working
properly, except for the Zeeman-setting results.

The problem is that when I tune the audio source around 53.53 kHz, per
both the manual and the sticker on the door, I don't see any change in
Beam I at all.  Nor do I see anything at the alternate frequency of
42.82 kHz.

Instead, I see the expected three peaks -- primary with a smaller
secondary on either side -- at about 48.21 kHz, which doesn't show up
anywhere in the literature I've found.

Below, I've cut and pasted a years-old message from TVB and Corby that
explains the Zeeman frequencies.  I've measured the synthesizer output
and it's nominally 12.6317725 MHz, which per that message should
correspond to a 53.53 Zeeman.  Where 48.21 kHz comes from, I have no idea.

I'm using a Rigol arbitrary function generator locked to an external
reference as the audio source, in sine wave mode.  I know that's not the
cleanest device in the world, but the wave doesn't look too bad on my
scope and a counter indicates the frequency is what the dial says.  As I
adjust the audio amplitude, the beam current responds, and I see a peak
at around 500mV, which the 5061B manual says is correct.

Any ideas why I might be seeing this very off-the-wall result?  Could
distortion in the audio source cause something like this?  I'm more
inclined to blame technique or gremlins than the 5061B -- again,
external measurements indicate that the thing is tuned correctly and
operating properly, just having this goofy Zeeman response.

Thanks!

John

[time-nuts] Zeeman frequency and cesium tube interchange
Tom Van Baak
Fri Apr 22 13:32:35 EDT 2005

Hi Brian,

The SI second is defined for mean sea level and no
external fields.

If there were no magnetic field, in theory, a cesium
tube would show a resonance when its synthesizer
generated exactly 9192.631770 MHz; the definition
of the second.

But in practice, a weak uniform magnetic field is
necessary for the beam apparatus to operate; to
isolate the center peak from the other peaks. This
DC field also has the side-effect of slightly shifting
the frequency of the center line.

Fortunately the frequency shift is a calculatable
amount (a function of magnetic field strength) so
the trick is that the synthesizer must be designed
to generate a slightly higher frequency to exactly
compensate for the shift that will be induced by
the field.

Thus the synthesizer for a 5061A does not actually
generate 9192 631 770 Hz as one might expect,
but because of the nominal 61 milligauss C-field,
the synthesizer must generate 9192 631 771.6 Hz
in order to lock onto the Cs peak precisely. The
Zeeman frequency for 61 mG is 42.82 kHz (info
from an old 5061A manual).

Other 5061A/B use a 76 mG field, corresponding
to a 53.53 kHz Zeeman frequency, and require the
synthesizer to generate 9192 631 772.5 Hz (info
from a new 5061B manual).

The short HP 5062C runs at 9192 631 774.3 Hz
with a Zeeman of 70.40 kHz.

Below is a great reply from Corby Dawson about
the problems this can cause when mixing FTS
4050, 4060, HP 5060A, 5061A, 5061B parts.

/tvb

----- Original Message -----

Tom,

The zeeman frequency required depends on two things, the magnitude of the
C-field current and the synthesizer frequency.

HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are
configured for a lower value of C-field current by selecting a higher
value series resistor on the A15 board. In this case any tube installed
(5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency.

HP 5061A and B units and 5060A units that have a synthesizer frequency of
12.6317716 Mhz are configured for a higher C-field current due to a lower
value series resistor in the C-field circuit. In this case any installed
tube will operate with a 42.82Khz zeeman frequency.

Problems arise when synthesizer and or A15 modules are swapped around
indiscriminantly leaving a unit with modules that do not match!

Since the HP and FTS tubes C-field windings are designed to provide the
same field for a given input they are interchangeable. You can operate
any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick
with how the mainframe came configured.

I don't remember what I ended up with as far as the zeeman freq. was
concerned when I installed a Frequency Electronics tube into a 5061A, but
do remember the line width was quite broad in keeping with the reduced
accuracy spec. of the FE tubes. (Same spec. as the 5062C tube)

I have installed FTS tubes into the 5062C giving it the +-7Z10-12th
accuracy spec. and did have to modify the synthesizer for 42.82Khz and
set the C-field accordingly.

Hope this helps!

Corby_


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

John, Did you do the Zeeman frequency adjustment before doing the 'degauss'? In other words, did you 'magnetize' rather than 'de-magnetize'? Have you checked the A15 board to make sure it has the correct resistor and current? I have a 5061B Service Manual if you need some sections or schematics. Good luck. Joe -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of John Ackermann N8UR Sent: Saturday, October 27, 2012 2:16 PM To: Discussion of precise time and frequency measurement Subject: [time-nuts] Zeeman frequency oddness Since I recently got an HP Cs degausser (thanks, Stijn!), I though I'd go through the whole setup routine for my 5061B/004 and see how close the C-field-via-Zeeman setting would bring me to GPS-derived frequency. It turned out to be an interesting and puzzling exercise. This 5061 seems to be in perfect working order -- quick lock, good meter readings, and measurements indicate frequency within parts in e12. I don't have any reason to believe that it's not tuned or working properly, except for the Zeeman-setting results. The problem is that when I tune the audio source around 53.53 kHz, per both the manual and the sticker on the door, I don't see any change in Beam I at all. Nor do I see anything at the alternate frequency of 42.82 kHz. Instead, I see the expected three peaks -- primary with a smaller secondary on either side -- at about 48.21 kHz, which doesn't show up anywhere in the literature I've found. Below, I've cut and pasted a years-old message from TVB and Corby that explains the Zeeman frequencies. I've measured the synthesizer output and it's nominally 12.6317725 MHz, which per that message should correspond to a 53.53 Zeeman. Where 48.21 kHz comes from, I have no idea. I'm using a Rigol arbitrary function generator locked to an external reference as the audio source, in sine wave mode. I know that's not the cleanest device in the world, but the wave doesn't look too bad on my scope and a counter indicates the frequency is what the dial says. As I adjust the audio amplitude, the beam current responds, and I see a peak at around 500mV, which the 5061B manual says is correct. Any ideas why I might be seeing this very off-the-wall result? Could distortion in the audio source cause something like this? I'm more inclined to blame technique or gremlins than the 5061B -- again, external measurements indicate that the thing is tuned correctly and operating properly, just having this goofy Zeeman response. Thanks! John [time-nuts] Zeeman frequency and cesium tube interchange Tom Van Baak Fri Apr 22 13:32:35 EDT 2005 Hi Brian, The SI second is defined for mean sea level and no external fields. If there were no magnetic field, in theory, a cesium tube would show a resonance when its synthesizer generated exactly 9192.631770 MHz; the definition of the second. But in practice, a weak uniform magnetic field is necessary for the beam apparatus to operate; to isolate the center peak from the other peaks. This DC field also has the side-effect of slightly shifting the frequency of the center line. Fortunately the frequency shift is a calculatable amount (a function of magnetic field strength) so the trick is that the synthesizer must be designed to generate a slightly higher frequency to exactly compensate for the shift that will be induced by the field. Thus the synthesizer for a 5061A does not actually generate 9192 631 770 Hz as one might expect, but because of the nominal 61 milligauss C-field, the synthesizer must generate 9192 631 771.6 Hz in order to lock onto the Cs peak precisely. The Zeeman frequency for 61 mG is 42.82 kHz (info from an old 5061A manual). Other 5061A/B use a 76 mG field, corresponding to a 53.53 kHz Zeeman frequency, and require the synthesizer to generate 9192 631 772.5 Hz (info from a new 5061B manual). The short HP 5062C runs at 9192 631 774.3 Hz with a Zeeman of 70.40 kHz. Below is a great reply from Corby Dawson about the problems this can cause when mixing FTS 4050, 4060, HP 5060A, 5061A, 5061B parts. /tvb ----- Original Message ----- > Tom, > > The zeeman frequency required depends on two things, the magnitude of the > C-field current and the synthesizer frequency. > > HP 5061A and B units that have a synthesizer freq. of 12.6317725 Mhz are > configured for a lower value of C-field current by selecting a higher > value series resistor on the A15 board. In this case any tube installed > (5061A/B 5060 4050 4060) will operate at the 53.53Khz zeeman frequency. > > HP 5061A and B units and 5060A units that have a synthesizer frequency of > 12.6317716 Mhz are configured for a higher C-field current due to a lower > value series resistor in the C-field circuit. In this case any installed > tube will operate with a 42.82Khz zeeman frequency. > > Problems arise when synthesizer and or A15 modules are swapped around > indiscriminantly leaving a unit with modules that do not match! > > Since the HP and FTS tubes C-field windings are designed to provide the > same field for a given input they are interchangeable. You can operate > any of these tubes at a 42.82 or 53.53Khz zeeman. I usually just stick > with how the mainframe came configured. > > > I don't remember what I ended up with as far as the zeeman freq. was > concerned when I installed a Frequency Electronics tube into a 5061A, but > do remember the line width was quite broad in keeping with the reduced > accuracy spec. of the FE tubes. (Same spec. as the 5062C tube) > > I have installed FTS tubes into the 5062C giving it the +-7Z10-12th > accuracy spec. and did have to modify the synthesizer for 42.82Khz and > set the C-field accordingly. > > Hope this helps! > > Corby_ _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
MD
Magnus Danielson
Sun, Oct 28, 2012 12:07 AM

On 10/28/2012 01:13 AM, Chuck Harris wrote:

Two things come to mind:

  1. is this the high performance tube? Perhaps it has a different
    Zeeman frequency than the standard tube?
  2. a non HP replacement tube? FTS Cs beam tubes are way different
    Zeeman frequencies than HP.

A high performance tube should give you better signal to noise.

A non HP tube might give you a somewhat different C-field setting for
same center lock-in.

The physics of Cs-133 is still the same, and the RF-chain is still the
same. If not, someone has to teach me something new, or at least inform
me why the high performance tubes are so much different.

Cheers,
Magnus

On 10/28/2012 01:13 AM, Chuck Harris wrote: > Two things come to mind: > > 1) is this the high performance tube? Perhaps it has a different > Zeeman frequency than the standard tube? > 2) a non HP replacement tube? FTS Cs beam tubes are way different > Zeeman frequencies than HP. A high performance tube should give you better signal to noise. A non HP tube might give you a somewhat different C-field setting for same center lock-in. The physics of Cs-133 is still the same, and the RF-chain is still the same. If not, someone has to teach me something new, or at least inform me why the high performance tubes are so much different. Cheers, Magnus
JF
J. Forster
Sun, Oct 28, 2012 12:48 AM

Is there perhaps a hemisphere jumper for the sign of nulling fields?

-John

==========

On 10/28/2012 01:13 AM, Chuck Harris wrote:

Two things come to mind:

  1. is this the high performance tube? Perhaps it has a different
    Zeeman frequency than the standard tube?
  2. a non HP replacement tube? FTS Cs beam tubes are way different
    Zeeman frequencies than HP.

A high performance tube should give you better signal to noise.

A non HP tube might give you a somewhat different C-field setting for
same center lock-in.

The physics of Cs-133 is still the same, and the RF-chain is still the
same. If not, someone has to teach me something new, or at least inform
me why the high performance tubes are so much different.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Is there perhaps a hemisphere jumper for the sign of nulling fields? -John ========== > On 10/28/2012 01:13 AM, Chuck Harris wrote: >> Two things come to mind: >> >> 1) is this the high performance tube? Perhaps it has a different >> Zeeman frequency than the standard tube? >> 2) a non HP replacement tube? FTS Cs beam tubes are way different >> Zeeman frequencies than HP. > > A high performance tube should give you better signal to noise. > > A non HP tube might give you a somewhat different C-field setting for > same center lock-in. > > The physics of Cs-133 is still the same, and the RF-chain is still the > same. If not, someone has to teach me something new, or at least inform > me why the high performance tubes are so much different. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >
CH
Chuck Harris
Sun, Oct 28, 2012 5:25 AM

Magnus Danielson wrote:

On 10/28/2012 01:13 AM, Chuck Harris wrote:

Two things come to mind:

  1. is this the high performance tube? Perhaps it has a different
    Zeeman frequency than the standard tube?
  2. a non HP replacement tube? FTS Cs beam tubes are way different
    Zeeman frequencies than HP.

A high performance tube should give you better signal to noise.

A non HP tube might give you a somewhat different C-field setting for same center
lock-in.

The FTS tubes in my 4050 have a Zeeman frequency of around 43KHz, which if
my estimate is right is about 10KHz different from the HP Zeeman frequency
of around 53KHz.  And as I recall, the little HP submarine C-Beam has yet
another Zeeman frequency, but I can't dig up its manual at the moment.

AFAIK, the Zeeman frequency is very much dependent on the architecture of
the particular physics package.

-Chuck Harris

The physics of Cs-133 is still the same, and the RF-chain is still the same. If not,
someone has to teach me something new, or at least inform me why the high performance
tubes are so much different.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Magnus Danielson wrote: > On 10/28/2012 01:13 AM, Chuck Harris wrote: >> Two things come to mind: >> >> 1) is this the high performance tube? Perhaps it has a different >> Zeeman frequency than the standard tube? >> 2) a non HP replacement tube? FTS Cs beam tubes are way different >> Zeeman frequencies than HP. > > A high performance tube should give you better signal to noise. > > A non HP tube might give you a somewhat different C-field setting for same center > lock-in. The FTS tubes in my 4050 have a Zeeman frequency of around 43KHz, which if my estimate is right is about 10KHz different from the HP Zeeman frequency of around 53KHz. And as I recall, the little HP submarine C-Beam has yet another Zeeman frequency, but I can't dig up its manual at the moment. AFAIK, the Zeeman frequency is very much dependent on the architecture of the particular physics package. -Chuck Harris > > The physics of Cs-133 is still the same, and the RF-chain is still the same. If not, > someone has to teach me something new, or at least inform me why the high performance > tubes are so much different. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
TV
Tom Van Baak (lab)
Sun, Oct 28, 2012 2:06 PM

Chuck,

Information about the HP submarine cesium standard (5062c) as well as details about Zeeman splitting are here:
http://www.leapsecond.com/museum/hp5062c/theory.htm

/tvb (iPhone4)

On Oct 28, 2012, at 1:25 AM, Chuck Harris cfharris@erols.com wrote:

Magnus Danielson wrote:

On 10/28/2012 01:13 AM, Chuck Harris wrote:

Two things come to mind:

  1. is this the high performance tube? Perhaps it has a different
    Zeeman frequency than the standard tube?
  2. a non HP replacement tube? FTS Cs beam tubes are way different
    Zeeman frequencies than HP.

A high performance tube should give you better signal to noise.

A non HP tube might give you a somewhat different C-field setting for same center
lock-in.

The FTS tubes in my 4050 have a Zeeman frequency of around 43KHz, which if
my estimate is right is about 10KHz different from the HP Zeeman frequency
of around 53KHz.  And as I recall, the little HP submarine C-Beam has yet
another Zeeman frequency, but I can't dig up its manual at the moment.

AFAIK, the Zeeman frequency is very much dependent on the architecture of
the particular physics package.

-Chuck Harris

The physics of Cs-133 is still the same, and the RF-chain is still the same. If not,
someone has to teach me something new, or at least inform me why the high performance
tubes are so much different.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Chuck, Information about the HP submarine cesium standard (5062c) as well as details about Zeeman splitting are here: http://www.leapsecond.com/museum/hp5062c/theory.htm /tvb (iPhone4) On Oct 28, 2012, at 1:25 AM, Chuck Harris <cfharris@erols.com> wrote: > Magnus Danielson wrote: >> On 10/28/2012 01:13 AM, Chuck Harris wrote: >>> Two things come to mind: >>> >>> 1) is this the high performance tube? Perhaps it has a different >>> Zeeman frequency than the standard tube? >>> 2) a non HP replacement tube? FTS Cs beam tubes are way different >>> Zeeman frequencies than HP. >> >> A high performance tube should give you better signal to noise. >> >> A non HP tube might give you a somewhat different C-field setting for same center >> lock-in. > > The FTS tubes in my 4050 have a Zeeman frequency of around 43KHz, which if > my estimate is right is about 10KHz different from the HP Zeeman frequency > of around 53KHz. And as I recall, the little HP submarine C-Beam has yet > another Zeeman frequency, but I can't dig up its manual at the moment. > > AFAIK, the Zeeman frequency is very much dependent on the architecture of > the particular physics package. > > -Chuck Harris > >> >> The physics of Cs-133 is still the same, and the RF-chain is still the same. If not, >> someone has to teach me something new, or at least inform me why the high performance >> tubes are so much different. >> >> Cheers, >> Magnus >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
TV
Tom Van Baak (lab)
Sun, Oct 28, 2012 2:25 PM

Cs tube ADEV is related to S/N ratio which is dependent on beam flux (the formula is in any quantum physics textbook).To achieve higher beam currents high-perf tubes run the oven at a higher temp (which is why their lifetime is much shorter).

So yes the RF chain is the same. But the oven controller set point is different and the E/M may be configured differently to accommodate the higher signal levels.

/tvb (iPhone4)

On Oct 27, 2012, at 8:07 PM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

On 10/28/2012 01:13 AM, Chuck Harris wrote:

Two things come to mind:

  1. is this the high performance tube? Perhaps it has a different
    Zeeman frequency than the standard tube?
  2. a non HP replacement tube? FTS Cs beam tubes are way different
    Zeeman frequencies than HP.

A high performance tube should give you better signal to noise.

A non HP tube might give you a somewhat different C-field setting for same center lock-in.

The physics of Cs-133 is still the same, and the RF-chain is still the same. If not, someone has to teach me something new, or at least inform me why the high performance tubes are so much different.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Cs tube ADEV is related to S/N ratio which is dependent on beam flux (the formula is in any quantum physics textbook).To achieve higher beam currents high-perf tubes run the oven at a higher temp (which is why their lifetime is much shorter). So yes the RF chain is the same. But the oven controller set point is different and the E/M may be configured differently to accommodate the higher signal levels. /tvb (iPhone4) On Oct 27, 2012, at 8:07 PM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > On 10/28/2012 01:13 AM, Chuck Harris wrote: >> Two things come to mind: >> >> 1) is this the high performance tube? Perhaps it has a different >> Zeeman frequency than the standard tube? >> 2) a non HP replacement tube? FTS Cs beam tubes are way different >> Zeeman frequencies than HP. > > A high performance tube should give you better signal to noise. > > A non HP tube might give you a somewhat different C-field setting for same center lock-in. > > The physics of Cs-133 is still the same, and the RF-chain is still the same. If not, someone has to teach me something new, or at least inform me why the high performance tubes are so much different. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.