Re: [volt-nuts] Traveling Standards

CP
Charles P. Steinmetz
Thu, Aug 25, 2011 10:57 PM

Charlie wrote:

I have some laboratory grade voltage standards here in my home
lab.  I have a Fluke 5440B direct voltage calibrator which is
cross-checked and compared against a pair of Fluke 732A DC Reference
Standards that run on a UPS and are kept hot all the time.  My
ultimate transfer uncertainty at 10 volts should be less than +/-
6.0 ppm per year and less than +/- 1.5ppm per 90 days over a
temperature range of 18 degrees C to 28 degrees C based on the Fluke
732A specifications.

The only problem is that I will not ship these units because that is
bad for calibration, stability, and uncertainty.  However, something
could be arranged if you want to ship the unit to be calibrated to me.

My understanding of Bob's original proposal was that he wants to
build a stable reference of uncertain voltage, and send that to
someone (or several someones) who will render an opinion on its
actual voltage and send it back.  A well-calibrated 5440B and a null
voltmeter (Fluke 845A/AB/AR, HP419A) would work, as would a 732A or
equivalent and a precision voltage divider (Fluke 720A or equivalent).

Bob did mention having the recipient power the traveling reference up
-- if possible, it should be shipped "hot" under battery power so
there is no power cycle/retrace to worry about.

Best regards,

Charles

Charlie wrote: >I have some laboratory grade voltage standards here in my home >lab. I have a Fluke 5440B direct voltage calibrator which is >cross-checked and compared against a pair of Fluke 732A DC Reference >Standards that run on a UPS and are kept hot all the time. My >ultimate transfer uncertainty at 10 volts should be less than +/- >6.0 ppm per year and less than +/- 1.5ppm per 90 days over a >temperature range of 18 degrees C to 28 degrees C based on the Fluke >732A specifications. > >The only problem is that I will not ship these units because that is >bad for calibration, stability, and uncertainty. However, something >could be arranged if you want to ship the unit to be calibrated to me. My understanding of Bob's original proposal was that he wants to build a stable reference of uncertain voltage, and send that to someone (or several someones) who will render an opinion on its actual voltage and send it back. A well-calibrated 5440B and a null voltmeter (Fluke 845A/AB/AR, HP419A) would work, as would a 732A or equivalent and a precision voltage divider (Fluke 720A or equivalent). Bob did mention having the recipient power the traveling reference up -- if possible, it should be shipped "hot" under battery power so there is no power cycle/retrace to worry about. Best regards, Charles
BS
Bob Smither
Fri, Aug 26, 2011 2:39 PM

Charles P. Steinmetz wrote:

<snip>

My understanding of Bob's original proposal was that he wants to build a
stable reference of uncertain voltage, and send that to someone (or
several someones) who will render an opinion on its actual voltage and
send it back.

Correct.

<snip>

Bob did mention having the recipient power the traveling reference up --
if possible, it should be shipped "hot" under battery power so there is
no power cycle/retrace to worry about.

I can see the advantage of this, but as an example the lm199ah draws
about 15 mA at 25C.  If exposed to low temperatures during shipment it
will draw more.  Perhaps this could be reduced if the part is put in a
thermally insulated box.  A one week shipment (out and back), at 15 mA,
would require a 2600 mA-hr battery.  Certain not impossible, but
somewhat heavy.

With the goal of 10ppm overall uncertainty, will the thermal hysteresis
be a limiting factor if the unit is shipped un-powered?  How long should
it be powered at the remote lab before a measurement is made?

Best regards,

Bob Smither, Ph.D.                                Smither@C-C-I.com


---==
But when a long Train of Abuses and Usurpations, pursuing invariably
the same Object, evinces a Design to reduce them under absolute
Despotism, it is their Right, it is their Duty, to throw off such
Government, and to provide new Guards for their future Security.
-- Thomas Jefferson, Declaration of Independence, July 4, 1776


---==

Charles P. Steinmetz wrote: <snip> > My understanding of Bob's original proposal was that he wants to build a > stable reference of uncertain voltage, and send that to someone (or > several someones) who will render an opinion on its actual voltage and > send it back. Correct. <snip> > Bob did mention having the recipient power the traveling reference up -- > if possible, it should be shipped "hot" under battery power so there is > no power cycle/retrace to worry about. I can see the advantage of this, but as an example the lm199ah draws about 15 mA at 25C. If exposed to low temperatures during shipment it will draw more. Perhaps this could be reduced if the part is put in a thermally insulated box. A one week shipment (out and back), at 15 mA, would require a 2600 mA-hr battery. Certain not impossible, but somewhat heavy. With the goal of 10ppm overall uncertainty, will the thermal hysteresis be a limiting factor if the unit is shipped un-powered? How long should it be powered at the remote lab before a measurement is made? Best regards, -- Bob Smither, Ph.D. Smither@C-C-I.com ==================================================================== But when a long Train of Abuses and Usurpations, pursuing invariably the same Object, evinces a Design to reduce them under absolute Despotism, it is their Right, it is their Duty, to throw off such Government, and to provide new Guards for their future Security. -- Thomas Jefferson, Declaration of Independence, July 4, 1776 ====================================================================
AJ
Andreas Jahn
Fri, Aug 26, 2011 3:49 PM

I can see the advantage of this, but as an example the lm199ah draws
about 15 mA at 25C.  If exposed to low temperatures during shipment it
will draw more.  Perhaps this could be reduced if the part is put in a
thermally insulated box.  A one week shipment (out and back), at 15 mA,
would require a 2600 mA-hr battery.  Certain not impossible, but
somewhat heavy.

Practically I can power a LM399 with 8 NiMH cells (9,6V 2000mAH)
for about 2 Days before I need to recharge.
I tried better thermal insulation but this has no significant influence.

With the goal of 10ppm overall uncertainty, will the thermal hysteresis
be a limiting factor if the unit is shipped un-powered?  How long should
it be powered at the remote lab before a measurement is made?

On my LM399 with a 6,5 Digit instrument (10uV resolution)
you will not see any thermal hysteresis.
(except when comparing 2 of them in a higher resolution range).

To the diagram:
Y-axis is the output of my 5V  24 Bit ADC in mV. X-axis is time in minutes.
(integration time 1 minute so measurement noise is around 1uV at ADC-input)
The reference is attached to the ADC by a precision 2:1 divider (LTC1043)
Voltage before power off 3431.174 * 2 mV
1 minute power off
1 minute value out of diagram
about 15 minutes warm-up time
about 1.5 uV * 2 = 3uV Hysteresis.
2nd power off for one minute off gives additional shift of 0.5 * 2 = 1uV.

So I would recommend a power-up time of at least 1 hour or so.
The instrument which is used to measure the value will need this too.

With best regards

Andreas

> I can see the advantage of this, but as an example the lm199ah draws > about 15 mA at 25C. If exposed to low temperatures during shipment it > will draw more. Perhaps this could be reduced if the part is put in a > thermally insulated box. A one week shipment (out and back), at 15 mA, > would require a 2600 mA-hr battery. Certain not impossible, but > somewhat heavy. Practically I can power a LM399 with 8 NiMH cells (9,6V 2000mAH) for about 2 Days before I need to recharge. I tried better thermal insulation but this has no significant influence. > With the goal of 10ppm overall uncertainty, will the thermal hysteresis > be a limiting factor if the unit is shipped un-powered? How long should > it be powered at the remote lab before a measurement is made? On my LM399 with a 6,5 Digit instrument (10uV resolution) you will not see any thermal hysteresis. (except when comparing 2 of them in a higher resolution range). To the diagram: Y-axis is the output of my 5V 24 Bit ADC in mV. X-axis is time in minutes. (integration time 1 minute so measurement noise is around 1uV at ADC-input) The reference is attached to the ADC by a precision 2:1 divider (LTC1043) Voltage before power off 3431.174 * 2 mV 1 minute power off 1 minute value out of diagram about 15 minutes warm-up time about 1.5 uV * 2 = 3uV Hysteresis. 2nd power off for one minute off gives additional shift of 0.5 * 2 = 1uV. So I would recommend a power-up time of at least 1 hour or so. The instrument which is used to measure the value will need this too. With best regards Andreas
BS
Bob Smither
Fri, Aug 26, 2011 5:23 PM

Andreas Jahn wrote:

<snip>

Practically I can power a LM399 with 8 NiMH cells (9,6V 2000mAH)
for about 2 Days before I need to recharge.

Too short to mail it out and return.  I think a line powered unit makes
the most sense (especially in light of your following data).

I tried better thermal insulation but this has no significant influence.

Haven't tried it yet - thanks for the input.

<snip>

On my LM399 with a 6,5 Digit instrument (10uV resolution)
you will not see any thermal hysteresis.
(except when comparing 2 of them in a higher resolution range).

This is what I gather from the data sheet.

To the diagram: Y-axis is the output of my 5V  24 Bit ADC in mV. X-axis
is time in minutes.

Thanks for the data.  Looks encouraging.

So I would recommend a power-up time of at least 1 hour or so.
The instrument which is used to measure the value will need this too.

Noted. Thanks.

Here are my back of the envelope preliminary design / calculations:

Power transformer / rectifier / capacitor => ~35 Volts DC.
LM7815A => 15 Volts (regulated).

LM199A: Vout =~ 7 V.
Set Id =~ 1 mA with an Rs = 8.06 K 1% resistor from 15 V to diode.
Power on-chip heater from 15 V.
Rout =~ 1 ohm => delta Is of 7 uA causes 1ppm shift.

Require Zin of measuring VM > 10M => 0.7 uA => 0.1 ppm error.

LM7815 TC =~ 1 mV/C => 0.125 uA/C => 0.018 ppm/C.
LM7815 line regulation =~ 13 mV/V.
For 35V +/- 10% => 46 mV => 5.7 uA => 0.813 ppm.

Rs TC =~ 100 ppm/C => dIs/dT =~ 100 ppm/C => 0.1 uA/C => 0.014 ppm/C.
Rs stability =~ .5% / 1000 hrs => 5.0 uA => 0.714 ppm.

Comments - feedback?

The two largest errors (other than 20 ppm stability of the lm199ah) are
both under 1 ppm.  They could be reduce further but is it worth it given
the stability of the lm199a?

Best regards,

Bob Smither

Andreas Jahn wrote: <snip> > Practically I can power a LM399 with 8 NiMH cells (9,6V 2000mAH) > for about 2 Days before I need to recharge. Too short to mail it out and return. I think a line powered unit makes the most sense (especially in light of your following data). > I tried better thermal insulation but this has no significant influence. Haven't tried it yet - thanks for the input. <snip> > On my LM399 with a 6,5 Digit instrument (10uV resolution) > you will not see any thermal hysteresis. > (except when comparing 2 of them in a higher resolution range). This is what I gather from the data sheet. > To the diagram: Y-axis is the output of my 5V 24 Bit ADC in mV. X-axis > is time in minutes. Thanks for the data. Looks encouraging. > So I would recommend a power-up time of at least 1 hour or so. > The instrument which is used to measure the value will need this too. Noted. Thanks. Here are my back of the envelope preliminary design / calculations: Power transformer / rectifier / capacitor => ~35 Volts DC. LM7815A => 15 Volts (regulated). LM199A: Vout =~ 7 V. Set Id =~ 1 mA with an Rs = 8.06 K 1% resistor from 15 V to diode. Power on-chip heater from 15 V. Rout =~ 1 ohm => delta Is of 7 uA causes 1ppm shift. Require Zin of measuring VM > 10M => 0.7 uA => 0.1 ppm error. LM7815 TC =~ 1 mV/C => 0.125 uA/C => 0.018 ppm/C. LM7815 line regulation =~ 13 mV/V. For 35V +/- 10% => 46 mV => 5.7 uA => 0.813 ppm. Rs TC =~ 100 ppm/C => dIs/dT =~ 100 ppm/C => 0.1 uA/C => 0.014 ppm/C. Rs stability =~ .5% / 1000 hrs => 5.0 uA => 0.714 ppm. Comments - feedback? The two largest errors (other than 20 ppm stability of the lm199ah) are both under 1 ppm. They could be reduce further but is it worth it given the stability of the lm199a? Best regards, Bob Smither
CP
Charles P. Steinmetz
Fri, Aug 26, 2011 5:44 PM

Bob wrote:

I can see the advantage of [shipping the reference under power], but
as an example the lm199ah draws about 15 mA at 25C.  If exposed to
low temperatures during shipment it will draw more.  Perhaps this
could be reduced if the part is put in a thermally insulated box.  A
one week shipment (out and back), at 15 mA, would require a 2600
mA-hr battery.  Certain not impossible, but somewhat heavy.

With the goal of 10ppm overall uncertainty, will the thermal
hysteresis be a limiting factor if the unit is shipped
un-powered?  How long should it be powered at the remote lab before
a measurement is made?

When one is trying to achieve repeatability in the single-digit ppm
range, I would not leave anything to chance -- and retrace
(hysteresis) is always a matter of chance, particularly if the device
has been powered off for a long time (several days) and may have been
exposed to significant temperature shifts during that time.  In my
view, the best approach in this instance is to use something like a
5Ah gel cell (to ensure relatively little voltage droop during the
transit time), with at least one stage of preregulation.  The total
package should be under 2 kg and should easily fit into a USPS
medium-size flat-rate box that ships for $11 (this assumes that you
and the recipient are both in the US).  UPS should be similar.  (If
$22 for a round trip is a deal-breaker, we might conclude that one
wasn't very serious about the project.)

You also might include a charger, so that the recipient can top off
the battery before measurement (if necessary) and before return
shipment.  All measurements would be made with the charger disconnected.
Best regards,

Charles

Bob wrote: >I can see the advantage of [shipping the reference under power], but >as an example the lm199ah draws about 15 mA at 25C. If exposed to >low temperatures during shipment it will draw more. Perhaps this >could be reduced if the part is put in a thermally insulated box. A >one week shipment (out and back), at 15 mA, would require a 2600 >mA-hr battery. Certain not impossible, but somewhat heavy. > >With the goal of 10ppm overall uncertainty, will the thermal >hysteresis be a limiting factor if the unit is shipped >un-powered? How long should it be powered at the remote lab before >a measurement is made? When one is trying to achieve repeatability in the single-digit ppm range, I would not leave anything to chance -- and retrace (hysteresis) is always a matter of chance, particularly if the device has been powered off for a long time (several days) and may have been exposed to significant temperature shifts during that time. In my view, the best approach in this instance is to use something like a 5Ah gel cell (to ensure relatively little voltage droop during the transit time), with at least one stage of preregulation. The total package should be under 2 kg and should easily fit into a USPS medium-size flat-rate box that ships for $11 (this assumes that you and the recipient are both in the US). UPS should be similar. (If $22 for a round trip is a deal-breaker, we might conclude that one wasn't very serious about the project.) You also might include a charger, so that the recipient can top off the battery before measurement (if necessary) and before return shipment. All measurements would be made with the charger disconnected. Best regards, Charles
BC
Brooke Clarke
Fri, Aug 26, 2011 5:48 PM

Hi:

Batteries are considered hazardous material and can not be shipped
without special Hazmat paper work which complicates things.  I think
there is an exception for some carriers, but not others, when the energy
is below a couple hundred watt hours (amp hours times volts).  The
battery pack  mentioned (2 AH * 9.6V = 19.2 WH).

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.End2PartyGovernment.com/

Bob Smither wrote:

Andreas Jahn wrote:

<snip>

Practically I can power a LM399 with 8 NiMH cells (9,6V 2000mAH)
for about 2 Days before I need to recharge.

Too short to mail it out and return.  I think a line powered unit
makes the most sense (especially in light of your following data).

I tried better thermal insulation but this has no significant influence.

Haven't tried it yet - thanks for the input.

<snip>

On my LM399 with a 6,5 Digit instrument (10uV resolution)
you will not see any thermal hysteresis.
(except when comparing 2 of them in a higher resolution range).

This is what I gather from the data sheet.

To the diagram: Y-axis is the output of my 5V  24 Bit ADC in mV.
X-axis is time in minutes.

Thanks for the data.  Looks encouraging.

So I would recommend a power-up time of at least 1 hour or so.
The instrument which is used to measure the value will need this too.

Noted. Thanks.

Here are my back of the envelope preliminary design / calculations:

Power transformer / rectifier / capacitor => ~35 Volts DC.
LM7815A => 15 Volts (regulated).

LM199A: Vout =~ 7 V.
Set Id =~ 1 mA with an Rs = 8.06 K 1% resistor from 15 V to diode.
Power on-chip heater from 15 V.
Rout =~ 1 ohm => delta Is of 7 uA causes 1ppm shift.

Require Zin of measuring VM > 10M => 0.7 uA => 0.1 ppm error.

LM7815 TC =~ 1 mV/C => 0.125 uA/C => 0.018 ppm/C.
LM7815 line regulation =~ 13 mV/V.
For 35V +/- 10% => 46 mV => 5.7 uA => 0.813 ppm.

Rs TC =~ 100 ppm/C => dIs/dT =~ 100 ppm/C => 0.1 uA/C => 0.014 ppm/C.
Rs stability =~ .5% / 1000 hrs => 5.0 uA => 0.714 ppm.

Comments - feedback?

The two largest errors (other than 20 ppm stability of the lm199ah)
are both under 1 ppm.  They could be reduce further but is it worth it
given the stability of the lm199a?

Best regards,

Bob Smither


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi: Batteries are considered hazardous material and can not be shipped without special Hazmat paper work which complicates things. I think there is an exception for some carriers, but not others, when the energy is below a couple hundred watt hours (amp hours times volts). The battery pack mentioned (2 AH * 9.6V = 19.2 WH). Have Fun, Brooke Clarke http://www.PRC68.com http://www.End2PartyGovernment.com/ Bob Smither wrote: > Andreas Jahn wrote: > > <snip> > >> Practically I can power a LM399 with 8 NiMH cells (9,6V 2000mAH) >> for about 2 Days before I need to recharge. > > Too short to mail it out and return. I think a line powered unit > makes the most sense (especially in light of your following data). > >> I tried better thermal insulation but this has no significant influence. > > Haven't tried it yet - thanks for the input. > > <snip> > >> On my LM399 with a 6,5 Digit instrument (10uV resolution) >> you will not see any thermal hysteresis. >> (except when comparing 2 of them in a higher resolution range). > > This is what I gather from the data sheet. > >> To the diagram: Y-axis is the output of my 5V 24 Bit ADC in mV. >> X-axis is time in minutes. > > Thanks for the data. Looks encouraging. > >> So I would recommend a power-up time of at least 1 hour or so. >> The instrument which is used to measure the value will need this too. > > Noted. Thanks. > > Here are my back of the envelope preliminary design / calculations: > > Power transformer / rectifier / capacitor => ~35 Volts DC. > LM7815A => 15 Volts (regulated). > > LM199A: Vout =~ 7 V. > Set Id =~ 1 mA with an Rs = 8.06 K 1% resistor from 15 V to diode. > Power on-chip heater from 15 V. > Rout =~ 1 ohm => delta Is of 7 uA causes 1ppm shift. > > Require Zin of measuring VM > 10M => 0.7 uA => 0.1 ppm error. > > LM7815 TC =~ 1 mV/C => 0.125 uA/C => 0.018 ppm/C. > LM7815 line regulation =~ 13 mV/V. > For 35V +/- 10% => 46 mV => 5.7 uA => 0.813 ppm. > > Rs TC =~ 100 ppm/C => dIs/dT =~ 100 ppm/C => 0.1 uA/C => 0.014 ppm/C. > Rs stability =~ .5% / 1000 hrs => 5.0 uA => 0.714 ppm. > > Comments - feedback? > > The two largest errors (other than 20 ppm stability of the lm199ah) > are both under 1 ppm. They could be reduce further but is it worth it > given the stability of the lm199a? > > Best regards, > > Bob Smither > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
BS
Bob Smither
Fri, Aug 26, 2011 6:13 PM

Brooke Clarke wrote:

Hi:

Batteries are considered hazardous material and can not be shipped
without special Hazmat paper work which complicates things.  I think
there is an exception for some carriers, but not others, when the energy
is below a couple hundred watt hours (amp hours times volts).  The
battery pack  mentioned (2 AH * 9.6V = 19.2 WH).

I wondered about this.  Since 9/11 there is precious little we can do
without getting someone's permission 8^(.

Andreas's results imply that, at least at the 10 ppm level, that
hysteresis may not be an issue.  It is stretching my measurement
capabilities, but I will try some experiments after I have some data on
several potential Traveling Standards.

Thanks,


---=====
Bob Smither, PhD                                Circuit Concepts, Inc.
"Every election is a sort of advance auction sale of stolen goods."
-- Libertarian H.L. Mencken
Smither@C-C-I.Com          http://www.C-C-I.Com          281-331-2744


---=====

Brooke Clarke wrote: > Hi: > > Batteries are considered hazardous material and can not be shipped > without special Hazmat paper work which complicates things. I think > there is an exception for some carriers, but not others, when the energy > is below a couple hundred watt hours (amp hours times volts). The > battery pack mentioned (2 AH * 9.6V = 19.2 WH). I wondered about this. Since 9/11 there is precious little we can do without getting someone's permission 8^(. Andreas's results imply that, at least at the 10 ppm level, that hysteresis may not be an issue. It is stretching my measurement capabilities, but I will try some experiments after I have some data on several potential Traveling Standards. Thanks, -- ======================================================================= Bob Smither, PhD Circuit Concepts, Inc. "Every election is a sort of advance auction sale of stolen goods." -- Libertarian H.L. Mencken Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744 =======================================================================
JD
John Devereux
Fri, Aug 26, 2011 6:27 PM

"Charles P. Steinmetz" charles_steinmetz@lavabit.com writes:

Bob wrote:

I can see the advantage of [shipping the reference under power], but
as an example the lm199ah draws about 15 mA at 25C.  If exposed to
low temperatures during shipment it will draw more.  Perhaps this
could be reduced if the part is put in a thermally insulated box.  A
one week shipment (out and back), at 15 mA, would require a 2600
mA-hr battery.  Certain not impossible, but somewhat heavy.

With the goal of 10ppm overall uncertainty, will the thermal
hysteresis be a limiting factor if the unit is shipped un-powered?
How long should it be powered at the remote lab before a measurement
is made?

When one is trying to achieve repeatability in the single-digit ppm
range, I would not leave anything to chance -- and retrace
(hysteresis) is always a matter of chance, particularly if the device
has been powered off for a long time (several days) and may have been
exposed to significant temperature shifts during that time.  In my
view, the best approach in this instance is to use something like a
5Ah gel cell (to ensure relatively little voltage droop during the
transit time), with at least one stage of preregulation.  The total
package should be under 2 kg and should easily fit into a USPS
medium-size flat-rate box that ships for $11 (this assumes that you
and the recipient are both in the US).  UPS should be similar.  (If
$22 for a round trip is a deal-breaker, we might conclude that one
wasn't very serious about the project.)

[...]

Hi Charles,

I think you will end up on Fox news and the Reference will end up
"defused" with a controlled explosion :)

I am still thinking LTZ1000 is the way to go on balance. It has all the
complications Andreas pointed out. But I think:

  • stable without keeping it always powered up

  • stable with RH

  • stable with ambient temperature (compared to others)

  • stable with respect to orientation, less likely to be affected by
    handling(?)

Does anyone know if the ebay / Chinese LTZ1000 are the real thing?
E.g. 250823975393.

Regards

John

--

John Devereux

"Charles P. Steinmetz" <charles_steinmetz@lavabit.com> writes: > Bob wrote: > >> I can see the advantage of [shipping the reference under power], but >> as an example the lm199ah draws about 15 mA at 25C. If exposed to >> low temperatures during shipment it will draw more. Perhaps this >> could be reduced if the part is put in a thermally insulated box. A >> one week shipment (out and back), at 15 mA, would require a 2600 >> mA-hr battery. Certain not impossible, but somewhat heavy. >> >> With the goal of 10ppm overall uncertainty, will the thermal >> hysteresis be a limiting factor if the unit is shipped un-powered? >> How long should it be powered at the remote lab before a measurement >> is made? > > When one is trying to achieve repeatability in the single-digit ppm > range, I would not leave anything to chance -- and retrace > (hysteresis) is always a matter of chance, particularly if the device > has been powered off for a long time (several days) and may have been > exposed to significant temperature shifts during that time. In my > view, the best approach in this instance is to use something like a > 5Ah gel cell (to ensure relatively little voltage droop during the > transit time), with at least one stage of preregulation. The total > package should be under 2 kg and should easily fit into a USPS > medium-size flat-rate box that ships for $11 (this assumes that you > and the recipient are both in the US). UPS should be similar. (If > $22 for a round trip is a deal-breaker, we might conclude that one > wasn't very serious about the project.) [...] Hi Charles, I think you will end up on Fox news and the Reference will end up "defused" with a controlled explosion :) I am still thinking LTZ1000 is the way to go on balance. It has all the complications Andreas pointed out. But I think: - stable without keeping it always powered up - stable with RH - stable with ambient temperature (compared to others) - stable with respect to orientation, less likely to be affected by handling(?) Does anyone know if the ebay / Chinese LTZ1000 are the real thing? E.g. 250823975393. Regards John -- John Devereux
CP
Charles P. Steinmetz
Fri, Aug 26, 2011 6:37 PM

Bob wrote:

Power transformer / rectifier / capacitor => ~35 Volts DC.
LM7815A => 15 Volts (regulated).

With 20 volts of overhead, why not use two stages of pre-regulation
(say, LM7824/LM7815) to improve the line regulation even
further?  You might also use separate LM7815s for the reference and
heater supplies, to minimize the reference supply changes due to
varying heater current and the associated load regulation/wiring drop.

The two largest errors (other than 20 ppm stability of the lm199ah)
are both under 1 ppm.  They could be reduce further but is it worth
it given the stability of the lm199a?

It is good design practice to reduce all sources of error to the
degree that they can can be reduced for little cost.  In this
context, I consider two additional three-terminal regulators "little
cost."  (If two additional regulators are just too much to bear, the
heater could be run from the 24 V supply.)

BTW, here is a standard similar to what you are
proposing:  http://gellerlabs.com/LNVR%20Series.htm  It uses an LM399
and the scaling amplifier of an AD587 to produce 10 V (I have
previously criticized this design for using the 587 instead of a
discrete op-amp that would give better performance, but the 587 used
this way appears to meet your design goals).  Geller also offers a
$40 bare board based on the 587 alone, for which he claims short-term
transfer accuracy of +/- 5
ppm:  http://gellerlabs.com/SVR%20Series.htm    Here is another one
(+/- 25 ppm at 5
V):  http://www.voltagestandard.com/New_Products.html    I believe
that both vendors will recalibrate them for a nominal fee.

Best regards,

Charles

Bob wrote: >Power transformer / rectifier / capacitor => ~35 Volts DC. >LM7815A => 15 Volts (regulated). With 20 volts of overhead, why not use two stages of pre-regulation (say, LM7824/LM7815) to improve the line regulation even further? You might also use separate LM7815s for the reference and heater supplies, to minimize the reference supply changes due to varying heater current and the associated load regulation/wiring drop. >The two largest errors (other than 20 ppm stability of the lm199ah) >are both under 1 ppm. They could be reduce further but is it worth >it given the stability of the lm199a? It is good design practice to reduce all sources of error to the degree that they can can be reduced for little cost. In this context, I consider two additional three-terminal regulators "little cost." (If two additional regulators are just too much to bear, the heater could be run from the 24 V supply.) BTW, here is a standard similar to what you are proposing: http://gellerlabs.com/LNVR%20Series.htm It uses an LM399 and the scaling amplifier of an AD587 to produce 10 V (I have previously criticized this design for using the 587 instead of a discrete op-amp that would give better performance, but the 587 used this way appears to meet your design goals). Geller also offers a $40 bare board based on the 587 alone, for which he claims short-term transfer accuracy of +/- 5 ppm: http://gellerlabs.com/SVR%20Series.htm Here is another one (+/- 25 ppm at 5 V): http://www.voltagestandard.com/New_Products.html I believe that both vendors will recalibrate them for a nominal fee. Best regards, Charles
CP
Charles P. Steinmetz
Fri, Aug 26, 2011 6:53 PM

John wrote:

I think you will end up on Fox news and the Reference will end up
"defused" with a controlled explosion :)

I recently ordered several of these, and I also quite often order
larger SLA batteries for UPS replacements.  They all arrive by UPS,
USPS, or FedEx with no drama and no hazmat labels that I can
see.  (But I don't think they are shipped by air, if someone is
contemplating that.)

I am still thinking LTZ1000 is the way to go

I concur, since "better is always better."  But it does appear that a
399 can, without undue difficulty, meet the design specifications, at
least if it is kept powered on.  I still think hysteresis is a
significant worry if the off time extends to days, particularly if it
is subjected to wide temperature changes during shipment.

Best regards,

Charles

John wrote: >I think you will end up on Fox news and the Reference will end up >"defused" with a controlled explosion :) I recently ordered several of these, and I also quite often order larger SLA batteries for UPS replacements. They all arrive by UPS, USPS, or FedEx with no drama and no hazmat labels that I can see. (But I don't think they are shipped by air, if someone is contemplating that.) >I am still thinking LTZ1000 is the way to go I concur, since "better is always better." But it does appear that a 399 can, without undue difficulty, meet the design specifications, at least if it is kept powered on. I still think hysteresis is a significant worry if the off time extends to days, particularly if it is subjected to wide temperature changes during shipment. Best regards, Charles