On 05/08/2012 12:52 PM, Charles Black wrote:
Hi Bob,
What a great job you are doing by organizing this cooperative field
comparison! It sounds like a lot of work though!
Looking at the first lab photo brought back memories of where I started
with my new (to me) 3458A three years ago! I hope it is OK to offer
suggestions that will make measuring microvolt levels on a 3458A more
accurate . Similar techniques can be used on any meter though.
I have been periodically monitoring my Datron 4910 and 4912 10V
standards for about 3 years looking at measurement drifts mainly
caused by temperature. I have many hours of recorded data looking at
0.1 PPM resolution drifts in my system. In order to improve data
stability I had to change my measuring technique from a BNC cable with
dual banana adapters (what is shown in the photo) to twisted pair
copper wires or shielded twisted pair wires (22 or 24 gauge) to
minimize input thermal offsets and settling time. Tin plating on the
wires measures about the same offset voltage as copper so it hasn't
been a problem to use it so far. If I put a huge thermal mass on my
input terminals (plastic dual banana plug) it can take many minutes to
stabilize and even then air currents will keep my meter from
stabilizing as completely as it can. Today my meter has about -240nv
offset voltage from a copper wire short but I see -1.2 microvolt with a
shorted dual banana plug, even a gold plated one. I have a shielded
dual banana to BNC adapter that's even worse.
My 3458A is very temperature sensitive so I constantly monitor the
temperature right in front of the input jacks to 0.1 degree C. I try to
keep mine at 23 degrees and do frequent ACAL's and repeat the ACAL if
more the terminal temperature varies much over a degree from 23C. The
TEMP? is also very important to watch and could be used to predict
temperature offset error. The TEMP? creeps up over a few months due to
dust in the filter but it is easy to clean with a little tap water.
The 3458A has a variable resolution feature that can be a problem at
high resolutions. Full DC resolution is acheived at NPLC = 1000. Lower
NPLC levels can be used to speed up measurements if loss of accuracy is
not a concern.
My shop has distinct thermal stratification so I try to keep the test
unit I am measuring at the same height as the meter and close to the
meter when doing accurate work.
As far as the portable standard goes - it would be great if it had gold
plated dual banana spaced terminal posts similar to the 3458A posts for
best repeatability and flexibility but, of course, any jacks will work
if they have enough time to stabilize.
Charlie
Thanks Charlie - great suggestions.
My goal has always been to know the values from the two references in the TS to
with 10 ppm. For the two refs (nominal 7 Volt and 5 Volt) this is 70uV and 50 uV.
I have one meter here that I can easily see thermal effects on (an old Fluke
845AR - 1uV full scale on most sensitive setting) so I appreciate your notes
about the use of copper wires.
Even though only trying for 10 ppm I appreciate that any source of error that
can be eliminated needs to be considered.
With your permission I would like to add your notes to the page here:
so I refer folks that are taking measurements on the TS to them.
---=======
Bob Smither bob@smither.net
If Tyranny and Oppression come to this land, it will be in the guise of
fighting a foreign enemy.
-James Madison
---=======
On 5/9/2012 11:53 AM, Bob Smither wrote:
On 05/08/2012 12:52 PM, Charles Black wrote:
Hi Bob,
What a great job you are doing by organizing this cooperative field
comparison! It sounds like a lot of work though!
Looking at the first lab photo brought back memories of where I started
with my new (to me) 3458A three years ago! I hope it is OK to offer
suggestions that will make measuring microvolt levels on a 3458A more
accurate . Similar techniques can be used on any meter though.
I have been periodically monitoring my Datron 4910 and 4912 10V
standards for about 3 years looking at measurement drifts mainly
caused by temperature. I have many hours of recorded data looking at
0.1 PPM resolution drifts in my system. In order to improve data
stability I had to change my measuring technique from a BNC cable with
dual banana adapters (what is shown in the photo) to twisted pair
copper wires or shielded twisted pair wires (22 or 24 gauge) to
minimize input thermal offsets and settling time. Tin plating on the
wires measures about the same offset voltage as copper so it hasn't
been a problem to use it so far. If I put a huge thermal mass on my
input terminals (plastic dual banana plug) it can take many minutes to
stabilize and even then air currents will keep my meter from
stabilizing as completely as it can. Today my meter has about -240nv
offset voltage from a copper wire short but I see -1.2 microvolt with a
shorted dual banana plug, even a gold plated one. I have a shielded
dual banana to BNC adapter that's even worse.
My 3458A is very temperature sensitive so I constantly monitor the
temperature right in front of the input jacks to 0.1 degree C. I try to
keep mine at 23 degrees and do frequent ACAL's and repeat the ACAL if
more the terminal temperature varies much over a degree from 23C. The
TEMP? is also very important to watch and could be used to predict
temperature offset error. The TEMP? creeps up over a few months due to
dust in the filter but it is easy to clean with a little tap water.
The 3458A has a variable resolution feature that can be a problem at
high resolutions. Full DC resolution is acheived at NPLC = 1000. Lower
NPLC levels can be used to speed up measurements if loss of accuracy is
not a concern.
My shop has distinct thermal stratification so I try to keep the test
unit I am measuring at the same height as the meter and close to the
meter when doing accurate work.
As far as the portable standard goes - it would be great if it had gold
plated dual banana spaced terminal posts similar to the 3458A posts for
best repeatability and flexibility but, of course, any jacks will work
if they have enough time to stabilize.
Charlie
Thanks Charlie - great suggestions.
My goal has always been to know the values from the two references in the TS to
with 10 ppm. For the two refs (nominal 7 Volt and 5 Volt) this is 70uV and 50 u
V.
My goal is to be to measure DC voltages between 10V and 1V with a 1 ppm
accuracy. I have my Datron 10V standard back from Fluke with each cell
at 0.6 uV/V measurement uncertainty (and data to 0.1 uV for 4 cells)
and a 3458A which has a 0.1 ppm linearity. The temperature drift of the
Datron 4910 is 0.05 ppm/DEGC. I checked the linearity of the 3458A with
my calibrated Fluke 720A and they agreed well within 0.1 ppm so my
3458A linearity uncertainty is 0.2 ppm or better. So far I have a total
uncertainty of 0.8 ppm and have to add in the transfer uncertainty
which can be as little as 0.02 ppm at 10V using my and an 845 similar
to yours. I can measure the Datron with the 3458A to see what the
temperature drift is and use that as a correction factor but I have
seen very little 3458A drift that is not temperature related. And I am
sure that there are other errors that need to be included as well.
I have one meter here that I can easily see thermal effects on (an old Fluke
845AR - 1uV full scale on most sensitive setting) so I appreciate your notes
about the use of copper wires.
Your first division is at 2uV on the 845's 100uV scale so you can
compare references within +/-100uV to 2uV. My 845AB wanders around
about +/-0.2uV so I usually use my HP419A which is stable. My 845 uses
the early input amp design and I am not sure how much drift is normal.
Even though only trying for 10 ppm I appreciate that any source of error that
can be eliminated needs to be considered.
With your permission I would like to add your notes to the page here:
[1]http://c-c-i.com/ts
so I refer folks that are taking measurements on the TS to them.
That's fine with me. You have my permission.
Best regards,
volt-nuts mailing list -- [2]volt-nuts@febo.com
To unsubscribe, go to [3]https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Charlie
References
The third volunteer to measure the TS has reported back his measurements and I
have posted them on the project page:
Thanks for the interest shown in this project! I have two more volunteers in
line to test the TS and will get it mailed to the next one this week.
A fellow volt-nut across the Atlantic has expressed interest, but we have not
found an economical way to ship it.
Best regards,
Bob Smither
I just dropped off the Volt-Nuts TS at my shipper, headed to the fourth
volunteer for this project. I will report his measurements as they become
available.
The prior readings are here:
--
Bob Smither, PhD Circuit Concepts, Inc.
---=======
Foreign aid might be defined as a transfer from poor people in
rich countries to rich people in poor countries.
--Douglas Casey (1992)
---=======
Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744(office) -4616(fax)
I added a heatsink and adjustable fanspeed control to my 7061. With the lid open at room temp 22 degrees the voltage meaured went up around 150 uV. In summer the meter get realy ot inside. In the test ( without cabinet around the pcb) the temp measured next to the ntc stayed 40 degrees plus/minus a half degree.
Voltage readings also increased in stabiltity.
http://www.pa4tim.nl/?p=2456 pictures are on the bottom of the page.
Fred PA4TIM
Op 19 sep. 2012 om 17:11 heeft Bob Smither smither@c-c-i.com het volgende geschreven:
The third volunteer to measure the TS has reported back his measurements and I
have posted them on the project page:
Thanks for the interest shown in this project! I have two more volunteers in
line to test the TS and will get it mailed to the next one this week.
A fellow volt-nut across the Atlantic has expressed interest, but we have not
found an economical way to ship it.
Best regards,
<smither.vcf>
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
On 09/22/2012 02:24 PM, Bob Smither wrote:
I just dropped off the Volt-Nuts TS at my shipper, headed to the fourth
volunteer for this project. I will report his measurements as they become
available.
Some drift data provided by the fourth volunteer is now available here:
--
"Of all tyrannies, a tyranny exercised for the good of its victims
may be the most oppressive. … [T]hose who torment us for our own
good will torment us without end, for they do so with the approval
of their own conscience."
-- C.S. Lewis
I have added more plots of the data taken by our fourth volunteer - so far over
a 3 week period. The plots are posted here:
Somewhat unexpected, at least by me, was the drift observed the first week and a
half. The two references (LM199A and MAX6350A) both drifted down several PPM.
The TS1 has been powered on for several thousands of hours - between being in my
lab and in the labs of other volunteers. It was on 24/7 for almost 2 months
here before being shipped to the fourth volunteer.
The TS1 is shipped un-powered.
There is a large humidity difference between my location (high humidity gulf
coast of Texas) and the location of the volunteer (high altitude, low humidity).
There is also a significant pressure difference between the two labs.
The LM199A is hermetically sealed.
Although a small sample, the two references appear to be similarly affected by
whatever caused the drift - similar range of drift, similar time constant.
Details (schematic, etc.) about the TS are available at the above link.
Any ideas about what could cause the drift we are seeing?
Thanks!
--
Bob Smither smither@c-c-i.com
---=======
In all criminal prosecutions, the accused shall enjoy the right to a
speedy and public trial, by an impartial jury of the State and district
wherein the crime shall have been committed; which district shall have
been previously ascertained by law, and to be informed of the nature and
cause of the accusation; to be confronted with the witnesses against him;
to have compulsory process for obtaining witnesses in his favor, and to
have the assistance of counsel for his defence.
-- Sixth amendment to the Constitution of the united States
---=======
Circuit Concepts, Inc. 281-331-2744
In message 507F6DC9.7030100@c-c-i.com, Bob Smither writes:
Any ideas about what could cause the drift we are seeing?
Something in the rest of the circuit ?
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
Hello Bob,
whats the matter with you. Infected by precision virus like me?
You wanted to have a standard with about 10ppm and now you blame a 3-4ppm
drift.
The LM199A is hermetically sealed.
The PCB, the 8K Resistor and the voltage Regulator are not.
Although a small sample, the two references appear to be similarly
affected by
whatever caused the drift - similar range of drift, similar time constant.
On the first view I would blame it on the meter.
It is very unusual that the drift of 2 different references has nearly exact
the same amount of ppm and direction.
But on the other side you state that there are several HP3458A which
recorded the drift.
It is not probable that all came freshly from calibration of a other
location.
Any ideas about what could cause the drift we are seeing?
From time constant it could be the humidity change.
My 2 LT1027CCN8-5 references which ara mounted only with 1 Pin
to the PCB have time constants in the range of 4-7 days.
The epoxy material of a PCB should lie in the same ball park area.
The LT1027 are influenced by around 0.5 ppm per percent humidity change.
For the hermetically sealed brand new references LT1236AILS they state in
their new product catalog
a humidity change of less than 10ppm for 25% humidity change. (page 36)
http://cds.linear.com/docs/Product%20Info/NPC.pdf
I asked them whether this is from mechanical stress from the PCB and they
confirmed to me
that with a dead bug mounting the influence of humidity will be virtually
unmeasurable.
So they will delete the parameter from the data sheet.
So for the LM399 it might be mechanical stress introduced by the PCB.
When looking at your cirquit there are several points to mention:
One common failure source will be the LM78L15. A output voltage change will
influence the supply of MAX6350 and the reference current of LM399.
PSRR of MAX6350 is about 2-5 ppm/V above 10V supply. (without self heating
effects).
The LM399 resistor will give a current change of about 10% per Volt (100uA)
resulting with 0.5 Ohm impedance in about 50uV/V or 7ppm/V
Other weak points of the cirquit are:
The LM399 heater voltage is not stabilized. this will give about 0.5ppm/V
And finally: was the LM399 always in the same orientation during
measurements?
(will be difficult with a cylindrical housing).
My LM399 drift 3-4 ppm by tilting orientation.
With best regards
Andreas
On 10/18/2012 02:05 PM, Andreas Jahn wrote:
Hello Bob,
whats the matter with you. Infected by precision virus like me?
You wanted to have a standard with about 10ppm and now you blame a 3-4ppm drift.
:-) - afraid so Andreas!
The LM199A is hermetically sealed.
The PCB, the 8K Resistor and the voltage Regulator are not.
Although a small sample, the two references appear to be similarly affected by
whatever caused the drift - similar range of drift, similar time constant.
On the first view I would blame it on the meter.
It is very unusual that the drift of 2 different references has nearly exact the
same amount of ppm and direction.
But on the other side you state that there are several HP3458A which recorded
the drift.
It is not probable that all came freshly from calibration of a other location.
Any ideas about what could cause the drift we are seeing?
From time constant it could be the humidity change.
This is my best guess.
My 2 LT1027CCN8-5 references which ara mounted only with 1 Pin
to the PCB have time constants in the range of 4-7 days.
The epoxy material of a PCB should lie in the same ball park area.
The LT1027 are influenced by around 0.5 ppm per percent humidity change.
For the hermetically sealed brand new references LT1236AILS they state in their
new product catalog
a humidity change of less than 10ppm for 25% humidity change. (page 36)
http://cds.linear.com/docs/Product%20Info/NPC.pdf
This is most interesting - so even "hermetically sealed" units are influenced by
humidity!
I asked them whether this is from mechanical stress from the PCB and they
confirmed to me
that with a dead bug mounting the influence of humidity will be virtually
unmeasurable.
So they will delete the parameter from the data sheet.
Mine are not "dead bug" mounted. The 'PCB' is in fact a Radio Shack perf board
So for the LM399 it might be mechanical stress introduced by the PCB.
When looking at your cirquit there are several points to mention:
One common failure source will be the LM78L15. A output voltage change will
influence the supply of MAX6350 and the reference current of LM399.
PSRR of MAX6350 is about 2-5 ppm/V above 10V supply. (without self heating
effects).
LM78L15 spec is 1mV/C. This would result in .02 ppm/C on the LM199A (operated
at 1 mA with 8K resistor providing the current).
For the MAX6350 1mV/C and 5 ppm/V => .005 ppm/C
The LM399 resistor will give a current change of about 10% per Volt (100uA)
resulting with 0.5 Ohm impedance in about 50uV/V or 7ppm/V
see note above.
Other weak points of the cirquit are:
The LM399 heater voltage is not stabilized. this will give about 0.5ppm/V
It is stabilized by the power supply - a 24 volt, .02%/C unit => 5mV/C => .0025
ppm/C.
And finally: was the LM399 always in the same orientation during measurements?
(will be difficult with a cylindrical housing).
My LM399 drift 3-4 ppm by tilting orientation.
Not sure - but from the consistent results (consistent drift and apparent
settling) it likely was.
The above still leaves humidity induced PCB changes causing mechanical stress as
likely. I am going to look for some better board material for when I construct TS2.
Thanks Andreas!
--
"As we act, let us not become the evil that we deplore."