Re: [time-nuts] HP5065A C-field current is temperature sensitive

T
timeok@timeok.it
Thu, Aug 20, 2015 5:38 PM

Hi ,

I think the Regulator Board should still be questioned.
To calibrate the +20 V,it is necessary to use an Extender Board because the calibration trimmer is not accessible with the card fitted normally.
I have used the extender HP built for service model 5060-0049. After alignment I could see that the +20 Volt  measured between the card A9 and A14(or other positions) have a variation of +50mV removing the extender.
I repeated the test several times with the same result.
The solution is to adjust the voltage with the extender to 19.95 volts and then remove the extender but a better solution will be to have a muti-turn trimmer accessible without removing the board.

Luciano

On Thu 20/08/15 10:55 , "Poul-Henning Kamp" phk@phk.freebsd.dk wrote:


In message ,
Bill Byrom writes:

I don't have an HP5065A and can only see the portion of the schematic
you copied.

The manual is on K04BB if you want the full monty.

Here is
what I can surmise from your measurements (assuming you don't have
any ground loops or measuring instrument or test lead thermoelectric

issues):

The measurement setup isn't optimal, I can only do one measurement
every five seconds on each point and the HP34972A is only a 6½ digit
instrument, but otherwise the setup is solid.

The change of the 20 V power supply when you lock is roughly 400
uV (20 ppm).

The change of the C-coil current isn't really repeatable. I see a big
500 nA (111 ppm) drop at 1,000 seconds, but no such clear change earlier
in the test.

It didn't actually lock until 1000 seconds, so one difference is that
the logic "continous operation" logic didn't reach final state and
its lamp didn't turn on in the previous attempts.

But the actual voltage changes you are measuring seem to be
roughly correlated between the 20 V power supply and C-coil current
sense resistor changes.

As they should be, because both are derived from the same A15CR5 zener.

I'm guessing about these values, based on converting your current
numbers into voltage based on a perfect R10 || R11 parallel combination
of 691.6 ohms. I see that R10 has an * asterisk, and I wonder what is
shown for that note. R10 might be a selected value at manufacturer, or
it might have a specific temperature coefficient.

The asterix means "selected". I have not been able to figure out
what criteria it is selected for.

The temperature coefficient of the resistors may be much more important
than anything else, especially for an old product.

All the important ones are wire-wound, probably for exactly that reason.

those old electrolytic capacitors (C4, C6, and C7 for example) are still
OK, or whether they are showing any changing leakage currents. You might
want to change them with new capacitors just in case.

Good point.

So I'm not convinced that the time curve is showing a correlation based
on the 20 V power supply affecting the C-coil current.

No, the main correlation is via the common zener, but the step at 1000s
is not present in the zener voltage, which means that the current
generator has really bad supply sensitivity.

It's possible that measuring system errors (such as where you connected

the

measurement system ground) might cause some of these changes. So you
might want to check your setup and the instrument and test lead
computed accuracy.

I think the way I've done it is OK. The 34972A has floating inputs
and I measure from a local GND for all six points. I'm not seing
any noise-artifacts.

I also disagree with your estimate of the CR5 zener current. [...]
[...]
the CR5 diode current must be (12.2195 - 4.2456 mA) = 7.9739 mA.

Good point.

That's
pretty close to the zero temperature coefficient current for the 1N938
you describe. So I do not recommend changing the diode current.

The 7.5 mA optimum is from a much later data-sheet, and may be for
a particular "high performance" variant of the 1N938, so there is
no guarantee that there even is a zero-tempco current for the one
in my HP5065.

Either way, fixing the zeners tempco is only half of the solution,
it looks to me like the "real" solution involves an entirely new
C-field current driver.

--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by
incompetence.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1]
and follow the instructions there.

Links:

[1]
http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
ilman/listinfo/time-nuts

Message sent via Atmail Open - http://atmail.org/

Hi , I think the Regulator Board should still be questioned. To calibrate the +20 V,it is necessary to use an Extender Board because the calibration trimmer is not accessible with the card fitted normally. I have used the extender HP built for service model 5060-0049. After alignment I could see that the +20 Volt measured between the card A9 and A14(or other positions) have a variation of +50mV removing the extender. I repeated the test several times with the same result. The solution is to adjust the voltage with the extender to 19.95 volts and then remove the extender but a better solution will be to have a muti-turn trimmer accessible without removing the board. Luciano On Thu 20/08/15 10:55 , "Poul-Henning Kamp" <phk@phk.freebsd.dk> wrote: > -------- > In message , > Bill Byrom writes: > > >I don't have an HP5065A and can only see the portion of the schematic > >you copied. > > The manual is on K04BB if you want the full monty. > > >Here is > >what I can surmise from your measurements (assuming you don't have > >any ground loops or measuring instrument or test lead thermoelectric > issues): > > The measurement setup isn't optimal, I can only do one measurement > every five seconds on each point and the HP34972A is only a 6½ digit > instrument, but otherwise the setup is solid. > > >The change of the 20 V power supply when you lock is roughly 400 > >uV (20 ppm). > > > >The change of the C-coil current isn't really repeatable. I see a big > >500 nA (111 ppm) drop at 1,000 seconds, but no such clear change earlier > >in the test. > > It didn't actually lock until 1000 seconds, so one difference is that > the logic "continous operation" logic didn't reach final state and > its lamp didn't turn on in the previous attempts. > > >But the actual voltage changes you are measuring seem to be > >roughly correlated between the 20 V power supply and C-coil current > >sense resistor changes. > > As they should be, because both are derived from the same A15CR5 zener. > > >I'm guessing about these values, based on converting your current > >numbers into voltage based on a perfect R10 || R11 parallel combination > >of 691.6 ohms. I see that R10 has an * asterisk, and I wonder what is > >shown for that note. R10 might be a selected value at manufacturer, or > >it might have a specific temperature coefficient. > > The asterix means "selected". I have not been able to figure out > what criteria it is selected for. > > >The temperature coefficient of the resistors may be much more important > >than anything else, especially for an old product. > > All the important ones are wire-wound, probably for exactly that reason. > > >those old electrolytic capacitors (C4, C6, and C7 for example) are still > >OK, or whether they are showing any changing leakage currents. You might > >want to change them with new capacitors just in case. > > Good point. > > >So I'm not convinced that the time curve is showing a correlation based > >on the 20 V power supply affecting the C-coil current. > > No, the main correlation is via the common zener, but the step at 1000s > is *not* present in the zener voltage, which means that the current > generator has really bad supply sensitivity. > > >It's possible that measuring system errors (such as where you connected > the > >measurement system ground) might cause some of these changes. So you > >might want to check your setup and the instrument and test lead > >computed accuracy. > > I think the way I've done it is OK. The 34972A has floating inputs > and I measure from a local GND for all six points. I'm not seing > any noise-artifacts. > > >I also disagree with your estimate of the CR5 zener current. [...] > >[...] > >the CR5 diode current must be (12.2195 - 4.2456 mA) = 7.9739 mA. > > Good point. > > >That's > >pretty close to the zero temperature coefficient current for the 1N938 > >you describe. So I do not recommend changing the diode current. > > The 7.5 mA optimum is from a much later data-sheet, and may be for > a particular "high performance" variant of the 1N938, so there is > no guarantee that there even is a zero-tempco current for the one > in my HP5065. > > Either way, fixing the zeners tempco is only half of the solution, > it looks to me like the "real" solution involves an entirely new > C-field current driver. > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by > incompetence. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1] > and follow the instructions there. > > > > Links: > ------ > [1] > http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma > ilman/listinfo/time-nuts > Message sent via Atmail Open - http://atmail.org/