W
WarrenS
Sat, Feb 6, 2010 7:19 PM
I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
Peat said:
>I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
>My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
Have fun
ws
*************
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
BC
Bob Camp
Sat, Feb 6, 2010 8:09 PM
Hi
It's possible / likely to injection lock with the tight loop approach and
get data that's much better than reality. A lot depends on the specific
oscillators under test and the buffers (if any) between the oscillators and
mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is
going to be at the 100 of nanovolts level. Certainly not impossible, but it
does present it's own set of issues. Lab gear to do it is available, but not
all that common. DC offsets and their temperature coefficients along with
thermocouple effects could make things exciting.
There is no perfect way to do any of this, only a lot of compromises here or
there. Each approach has stuff you need to watch out for.
Bob
From: "WarrenS" warrensjmail-one@yahoo.com
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] ADEV vs MDEV
I would appreciate any comments or observations on the topic of apparatus
with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability
measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference
Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer with a
Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end
test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it
takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that
allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec
(at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz
resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw
data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there,
'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
Hi
It's possible / likely to injection lock with the tight loop approach and
get data that's much better than reality. A lot depends on the specific
oscillators under test and the buffers (if any) between the oscillators and
mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is
going to be at the 100 of nanovolts level. Certainly not impossible, but it
does present it's own set of issues. Lab gear to do it is available, but not
all that common. DC offsets and their temperature coefficients along with
thermocouple effects could make things exciting.
There is no perfect way to do any of this, only a lot of compromises here or
there. Each approach has stuff you need to watch out for.
Bob
--------------------------------------------------
From: "WarrenS" <warrensjmail-one@yahoo.com>
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement"
<time-nuts@febo.com>
Subject: Re: [time-nuts] ADEV vs MDEV
>
> Peat said:
>>I would appreciate any comments or observations on the topic of apparatus
>>with demonstrated stability measurements.
>>My motivation is to discover the SIMPLEST scheme for making stability
>>measurements at the 1E-13 in 1s performance level.
>
>
> If you accept that the measurement is going to limited by the Reference
> Osc,
> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
> Method of measuring Freq stability".
> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>
>
> By replacing the "Voltage to freq converter, Freq counter & Printer with a
> Radio shack type PC data logging DVM,
> It can be up and running from scratch in under an Hr, with no high end
> test equipment needed.
> If you want performance that exceeds the best of most DMTD at low Tau it
> takes a little more work
> and a higher speed oversampling ADC data logger and a good offset voltage.
>
> I must add this is not a popular solution (Or a general Purpose one) but
> IF you know analog and have a GOOD osc with EFC to use for the reference,
> as far as I've been able to determine it is the BEST SIMPLE answer that
> allows High performance.
> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec
> (at 30 Hz Bandwidth)
>
> Basic modified NIST Block Diag attached:
> The NIST paper sums it up quite nicely:
> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz
> resolution
> so one has excellent precision capabilities with this system.'
>
> This does not address your other question of ADEV vs MDEV,
> What I've described is just a simple way to get the Low cost, GOOD Raw
> data.
> What you then do with that Data is a different subject.
>
> You can run the raw data thru one of the many ADEV programs out there,
> 'Plotter' being my choice.
>
>
> Have fun
> ws
>
> *************
>
> [time-nuts] ADEV vs MDEV
> Pete Rawson peterawson at earthlink.net
> Sat Feb 6 03:59:18 UTC 2010
>
> Efforts are underway to develop a low cost DMTD apparatus with
> demonstrated stability measurements of 1E-13 in 1s. It seems that
> existing TI counters can reach this goal in 10s. (using MDEV estimate
> or 100+s. using ADEV estimate). The question is; does the MDEV tool
> provide an appropriate measure of stability in this time range, or is
> the ADEV estimate a more correct answer?
>
> The TI performance I'm referring to is the 20-25 ps, single shot TI,
> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
> from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
> other counters behave similarly.
>
> I would appreciate any comments or observations on this topic.
> My motivation is to discover the simplest scheme for making
> stability measurements at this performance level; this is NOT
> even close to the state-of-the-art, but can still be useful.
>
> Pete Rawson
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
BG
Bruce Griffiths
Sat, Feb 6, 2010 8:11 PM
Sounds good but you still haven't found its Achilles heel:
The frequency measures need to be integrated (either implicitly or
explicitly) to produce phase measures which can then be used to
calculate ADEV, MDEV etc.
The major problem is that integration amplifies the small errors that
are inevitably present.
In practice (except for very noisy sources) the technique isnt
particularly useful for Tau more than a few times the inverse PLL bandwidth.
Thus NIST and others quietly dropped this method several decades ago.
This is alluded to in Steins recent paper availble on the Symmetricom
website:
The Allan Variance – Challenges and Opportunities
Bruce
WarrenS wrote:
I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter& Printer with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Sounds good but you still haven't found its Achilles heel:
The frequency measures need to be integrated (either implicitly or
explicitly) to produce phase measures which can then be used to
calculate ADEV, MDEV etc.
The major problem is that integration amplifies the small errors that
are inevitably present.
In practice (except for very noisy sources) the technique isnt
particularly useful for Tau more than a few times the inverse PLL bandwidth.
Thus NIST and others quietly dropped this method several decades ago.
This is alluded to in Steins recent paper availble on the Symmetricom
website:
*The Allan Variance – Challenges and Opportunities*
Bruce
WarrenS wrote:
> Peat said:
>
>> I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
>> My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
>>
>
> If you accept that the measurement is going to limited by the Reference Osc,
> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>
>
> By replacing the "Voltage to freq converter, Freq counter& Printer with a Radio shack type PC data logging DVM,
> It can be up and running from scratch in under an Hr, with no high end test equipment needed.
> If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
> and a higher speed oversampling ADC data logger and a good offset voltage.
>
> I must add this is not a popular solution (Or a general Purpose one) but
> IF you know analog and have a GOOD osc with EFC to use for the reference,
> as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
>
> Basic modified NIST Block Diag attached:
> The NIST paper sums it up quite nicely:
> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
> so one has excellent precision capabilities with this system.'
>
> This does not address your other question of ADEV vs MDEV,
> What I've described is just a simple way to get the Low cost, GOOD Raw data.
> What you then do with that Data is a different subject.
>
> You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
>
>
> Have fun
> ws
>
> *************
>
> [time-nuts] ADEV vs MDEV
> Pete Rawson peterawson at earthlink.net
> Sat Feb 6 03:59:18 UTC 2010
>
> Efforts are underway to develop a low cost DMTD apparatus with
> demonstrated stability measurements of 1E-13 in 1s. It seems that
> existing TI counters can reach this goal in 10s. (using MDEV estimate
> or 100+s. using ADEV estimate). The question is; does the MDEV tool
> provide an appropriate measure of stability in this time range, or is
> the ADEV estimate a more correct answer?
>
> The TI performance I'm referring to is the 20-25 ps, single shot TI,
> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
> from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
> other counters behave similarly.
>
> I would appreciate any comments or observations on this topic.
> My motivation is to discover the simplest scheme for making
> stability measurements at this performance level; this is NOT
> even close to the state-of-the-art, but can still be useful.
>
> Pete Rawson
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
BG
Bruce Griffiths
Sat, Feb 6, 2010 8:30 PM
To a first approximation injection locking alters the loop parameters so
its important to measure the actual PLL characteristics with the loop
closed and not just use the PLL parameters inferred from the OCXO EFC
transfer function etc.
The noise of the OCXO used as a VCXO will limit the noise floor.
An ADEV noise floor of 1E-13 isnt likely when using an HP10811A as the
VCXO for example.
Bruce
Bob Camp wrote:
Hi
It's possible / likely to injection lock with the tight loop approach
and get data that's much better than reality. A lot depends on the
specific oscillators under test and the buffers (if any) between the
oscillators and mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in
10^14 is going to be at the 100 of nanovolts level. Certainly not
impossible, but it does present it's own set of issues. Lab gear to do
it is available, but not all that common. DC offsets and their
temperature coefficients along with thermocouple effects could make
things exciting.
There is no perfect way to do any of this, only a lot of compromises
here or there. Each approach has stuff you need to watch out for.
Bob
From: "WarrenS" warrensjmail-one@yahoo.com
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] ADEV vs MDEV
I would appreciate any comments or observations on the topic of
apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making
stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the
Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that
level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer
with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high
end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau
it takes a little more work
and a higher speed oversampling ADC data logger and a good offset
voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the
reference,
as far as I've been able to determine it is the BEST SIMPLE answer
that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1
sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz
resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD
Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out
there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
To a first approximation injection locking alters the loop parameters so
its important to measure the actual PLL characteristics with the loop
closed and not just use the PLL parameters inferred from the OCXO EFC
transfer function etc.
The noise of the OCXO used as a VCXO will limit the noise floor.
An ADEV noise floor of 1E-13 isnt likely when using an HP10811A as the
VCXO for example.
Bruce
Bob Camp wrote:
> Hi
>
> It's possible / likely to injection lock with the tight loop approach
> and get data that's much better than reality. A lot depends on the
> specific oscillators under test and the buffers (if any) between the
> oscillators and mixer.
>
> If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in
> 10^14 is going to be at the 100 of nanovolts level. Certainly not
> impossible, but it does present it's own set of issues. Lab gear to do
> it is available, but not all that common. DC offsets and their
> temperature coefficients along with thermocouple effects could make
> things exciting.
>
> There is no perfect way to do any of this, only a lot of compromises
> here or there. Each approach has stuff you need to watch out for.
>
> Bob
>
> --------------------------------------------------
> From: "WarrenS" <warrensjmail-one@yahoo.com>
> Sent: Saturday, February 06, 2010 2:19 PM
> To: "Discussion of precise time and frequency measurement"
> <time-nuts@febo.com>
> Subject: Re: [time-nuts] ADEV vs MDEV
>
>>
>> Peat said:
>>> I would appreciate any comments or observations on the topic of
>>> apparatus with demonstrated stability measurements.
>>> My motivation is to discover the SIMPLEST scheme for making
>>> stability measurements at the 1E-13 in 1s performance level.
>>
>>
>> If you accept that the measurement is going to limited by the
>> Reference Osc,
>> for Low COST and SIMPLE, with the ability to measure ADEVs at that
>> level,
>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
>> Method of measuring Freq stability".
>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>
>>
>> By replacing the "Voltage to freq converter, Freq counter & Printer
>> with a Radio shack type PC data logging DVM,
>> It can be up and running from scratch in under an Hr, with no high
>> end test equipment needed.
>> If you want performance that exceeds the best of most DMTD at low Tau
>> it takes a little more work
>> and a higher speed oversampling ADC data logger and a good offset
>> voltage.
>>
>> I must add this is not a popular solution (Or a general Purpose one) but
>> IF you know analog and have a GOOD osc with EFC to use for the
>> reference,
>> as far as I've been able to determine it is the BEST SIMPLE answer
>> that allows High performance.
>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1
>> sec (at 30 Hz Bandwidth)
>>
>> Basic modified NIST Block Diag attached:
>> The NIST paper sums it up quite nicely:
>> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz
>> resolution
>> so one has excellent precision capabilities with this system.'
>>
>> This does not address your other question of ADEV vs MDEV,
>> What I've described is just a simple way to get the Low cost, GOOD
>> Raw data.
>> What you then do with that Data is a different subject.
>>
>> You can run the raw data thru one of the many ADEV programs out
>> there, 'Plotter' being my choice.
>>
>>
>> Have fun
>> ws
>>
>> *************
>>
>> [time-nuts] ADEV vs MDEV
>> Pete Rawson peterawson at earthlink.net
>> Sat Feb 6 03:59:18 UTC 2010
>>
>> Efforts are underway to develop a low cost DMTD apparatus with
>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>> provide an appropriate measure of stability in this time range, or is
>> the ADEV estimate a more correct answer?
>>
>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>> from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
>> other counters behave similarly.
>>
>> I would appreciate any comments or observations on this topic.
>> My motivation is to discover the simplest scheme for making
>> stability measurements at this performance level; this is NOT
>> even close to the state-of-the-art, but can still be useful.
>>
>> Pete Rawson
>>
>>
>
>
>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
BC
Bob Camp
Sat, Feb 6, 2010 8:43 PM
Hi
Unfortunately, the most common approach is to simply say "it's a wide band loop, no need to check it ...."
Bob
On Feb 6, 2010, at 3:30 PM, Bruce Griffiths wrote:
To a first approximation injection locking alters the loop parameters so its important to measure the actual PLL characteristics with the loop closed and not just use the PLL parameters inferred from the OCXO EFC transfer function etc.
The noise of the OCXO used as a VCXO will limit the noise floor.
An ADEV noise floor of 1E-13 isnt likely when using an HP10811A as the VCXO for example.
Bruce
Bob Camp wrote:
Hi
It's possible / likely to injection lock with the tight loop approach and get data that's much better than reality. A lot depends on the specific oscillators under test and the buffers (if any) between the oscillators and mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is going to be at the 100 of nanovolts level. Certainly not impossible, but it does present it's own set of issues. Lab gear to do it is available, but not all that common. DC offsets and their temperature coefficients along with thermocouple effects could make things exciting.
There is no perfect way to do any of this, only a lot of compromises here or there. Each approach has stuff you need to watch out for.
Bob
From: "WarrenS" warrensjmail-one@yahoo.com
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Subject: Re: [time-nuts] ADEV vs MDEV
I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
Hi
Unfortunately, the most common approach is to simply say "it's a wide band loop, no need to check it ...."
Bob
On Feb 6, 2010, at 3:30 PM, Bruce Griffiths wrote:
> To a first approximation injection locking alters the loop parameters so its important to measure the actual PLL characteristics with the loop closed and not just use the PLL parameters inferred from the OCXO EFC transfer function etc.
>
> The noise of the OCXO used as a VCXO will limit the noise floor.
> An ADEV noise floor of 1E-13 isnt likely when using an HP10811A as the VCXO for example.
>
> Bruce
>
> Bob Camp wrote:
>> Hi
>>
>> It's possible / likely to injection lock with the tight loop approach and get data that's much better than reality. A lot depends on the specific oscillators under test and the buffers (if any) between the oscillators and mixer.
>>
>> If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is going to be at the 100 of nanovolts level. Certainly not impossible, but it does present it's own set of issues. Lab gear to do it is available, but not all that common. DC offsets and their temperature coefficients along with thermocouple effects could make things exciting.
>>
>> There is no perfect way to do any of this, only a lot of compromises here or there. Each approach has stuff you need to watch out for.
>>
>> Bob
>>
>> --------------------------------------------------
>> From: "WarrenS" <warrensjmail-one@yahoo.com>
>> Sent: Saturday, February 06, 2010 2:19 PM
>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>> Subject: Re: [time-nuts] ADEV vs MDEV
>>
>>>
>>> Peat said:
>>>> I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
>>>> My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
>>>
>>>
>>> If you accept that the measurement is going to limited by the Reference Osc,
>>> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
>>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
>>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>>
>>>
>>> By replacing the "Voltage to freq converter, Freq counter & Printer with a Radio shack type PC data logging DVM,
>>> It can be up and running from scratch in under an Hr, with no high end test equipment needed.
>>> If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
>>> and a higher speed oversampling ADC data logger and a good offset voltage.
>>>
>>> I must add this is not a popular solution (Or a general Purpose one) but
>>> IF you know analog and have a GOOD osc with EFC to use for the reference,
>>> as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
>>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
>>>
>>> Basic modified NIST Block Diag attached:
>>> The NIST paper sums it up quite nicely:
>>> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
>>> so one has excellent precision capabilities with this system.'
>>>
>>> This does not address your other question of ADEV vs MDEV,
>>> What I've described is just a simple way to get the Low cost, GOOD Raw data.
>>> What you then do with that Data is a different subject.
>>>
>>> You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
>>>
>>>
>>> Have fun
>>> ws
>>>
>>> *************
>>>
>>> [time-nuts] ADEV vs MDEV
>>> Pete Rawson peterawson at earthlink.net
>>> Sat Feb 6 03:59:18 UTC 2010
>>>
>>> Efforts are underway to develop a low cost DMTD apparatus with
>>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>>> provide an appropriate measure of stability in this time range, or is
>>> the ADEV estimate a more correct answer?
>>>
>>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>>> from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
>>> other counters behave similarly.
>>>
>>> I would appreciate any comments or observations on this topic.
>>> My motivation is to discover the simplest scheme for making
>>> stability measurements at this performance level; this is NOT
>>> even close to the state-of-the-art, but can still be useful.
>>>
>>> Pete Rawson
>>>
>>>
>>
>>
>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
W
WarrenS
Sat, Feb 6, 2010 8:56 PM
Thus NIST and others quietly dropped this method several decades ago.
Could it be another reason?
I'll bet that was after they wanted to do better than 1e14 resolution AND
had unlimited amounts of time and Money,
Something most time Nuts are not blessed with. I Never said it was the BEST
way.
JUST given the goal, which was 1e13 in one second, there is not a simpler
and cheaper way to do it.
And nothing you said counter that point.
The frequency measures need to be integrated (either implicitly or
explicitly) to produce phase measures which can then be used to calculate
ADEV, MDEV etc.
Well ONE of us certainly has something backward.
To calculate ADEV, MDEV etc. YOU need Freq Differences.
The first thing that happens when phase is used is that it is turned into
Freq by taking the difference between each sample.
Integrated Freq data, which is what "Tight Phase-Lock Loop Method" gives you
directly (no Phase conversion needed),
Need not FIRST turned into Phase so that it can then be turned back into
Freq.
BUT in any case there is no difference in the noise, for a given bandwidth,
If you don't run out of digits and You have enough resolution.
The "Tight Phase-Lock Loop Method" can EASY get sub pS resolution, which is
better than most other ways.
AND don't need filters and slue rate control and multistage limiters and on
& on to do it, an RC works fine to replace all the stuff.
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Saturday, February 06, 2010 12:11 PM
Subject: Re: [time-nuts] ADEV vs MDEV
Sounds good but you still haven't found its Achilles heel:
The frequency measures need to be integrated (either implicitly or
explicitly) to produce phase measures which can then be used to calculate
ADEV, MDEV etc.
The major problem is that integration amplifies the small errors that are
inevitably present.
In practice (except for very noisy sources) the technique isnt
particularly useful for Tau more than a few times the inverse PLL
bandwidth.
Thus NIST and others quietly dropped this method several decades ago.
This is alluded to in Steins recent paper availble on the Symmetricom
website:
The Allan Variance – Challenges and Opportunities
Bruce
WarrenS wrote:
I would appreciate any comments or observations on the topic of
apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability
measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference
Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter& Printer with
a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end
test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it
takes a little more work
and a higher speed oversampling ADC data logger and a good offset
voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the
reference,
as far as I've been able to determine it is the BEST SIMPLE answer that
allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec
(at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz
resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw
data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there,
'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Bruce said:
> Thus NIST and others quietly dropped this method several decades ago.
Could it be another reason?
I'll bet that was after they wanted to do better than 1e14 resolution AND
had unlimited amounts of time and Money,
Something most time Nuts are not blessed with. I Never said it was the BEST
way.
JUST given the goal, which was 1e13 in one second, there is not a simpler
and cheaper way to do it.
And nothing you said counter that point.
> The frequency measures need to be integrated (either implicitly or
> explicitly) to produce phase measures which can then be used to calculate
> ADEV, MDEV etc.
Well ONE of us certainly has something backward.
To calculate ADEV, MDEV etc. YOU need Freq Differences.
The first thing that happens when phase is used is that it is turned into
Freq by taking the difference between each sample.
Integrated Freq data, which is what "Tight Phase-Lock Loop Method" gives you
directly (no Phase conversion needed),
Need not FIRST turned into Phase so that it can then be turned back into
Freq.
BUT in any case there is no difference in the noise, for a given bandwidth,
If you don't run out of digits and You have enough resolution.
The "Tight Phase-Lock Loop Method" can EASY get sub pS resolution, which is
better than most other ways.
AND don't need filters and slue rate control and multistage limiters and on
& on to do it, an RC works fine to replace all the stuff.
ws
*****************
----- Original Message -----
From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
To: "Discussion of precise time and frequency measurement"
<time-nuts@febo.com>
Sent: Saturday, February 06, 2010 12:11 PM
Subject: Re: [time-nuts] ADEV vs MDEV
> Sounds good but you still haven't found its Achilles heel:
>
> The frequency measures need to be integrated (either implicitly or
> explicitly) to produce phase measures which can then be used to calculate
> ADEV, MDEV etc.
> The major problem is that integration amplifies the small errors that are
> inevitably present.
> In practice (except for very noisy sources) the technique isnt
> particularly useful for Tau more than a few times the inverse PLL
> bandwidth.
>
> Thus NIST and others quietly dropped this method several decades ago.
> This is alluded to in Steins recent paper availble on the Symmetricom
> website:
>
> *The Allan Variance – Challenges and Opportunities*
>
>
> Bruce
>
> WarrenS wrote:
>> Peat said:
>>
>>> I would appreciate any comments or observations on the topic of
>>> apparatus with demonstrated stability measurements.
>>> My motivation is to discover the SIMPLEST scheme for making stability
>>> measurements at the 1E-13 in 1s performance level.
>>>
>>
>> If you accept that the measurement is going to limited by the Reference
>> Osc,
>> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
>> Method of measuring Freq stability".
>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>
>>
>> By replacing the "Voltage to freq converter, Freq counter& Printer with
>> a Radio shack type PC data logging DVM,
>> It can be up and running from scratch in under an Hr, with no high end
>> test equipment needed.
>> If you want performance that exceeds the best of most DMTD at low Tau it
>> takes a little more work
>> and a higher speed oversampling ADC data logger and a good offset
>> voltage.
>>
>> I must add this is not a popular solution (Or a general Purpose one) but
>> IF you know analog and have a GOOD osc with EFC to use for the
>> reference,
>> as far as I've been able to determine it is the BEST SIMPLE answer that
>> allows High performance.
>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec
>> (at 30 Hz Bandwidth)
>>
>> Basic modified NIST Block Diag attached:
>> The NIST paper sums it up quite nicely:
>> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz
>> resolution
>> so one has excellent precision capabilities with this system.'
>>
>> This does not address your other question of ADEV vs MDEV,
>> What I've described is just a simple way to get the Low cost, GOOD Raw
>> data.
>> What you then do with that Data is a different subject.
>>
>> You can run the raw data thru one of the many ADEV programs out there,
>> 'Plotter' being my choice.
>>
>>
>> Have fun
>> ws
>>
>> *************
>>
>> [time-nuts] ADEV vs MDEV
>> Pete Rawson peterawson at earthlink.net
>> Sat Feb 6 03:59:18 UTC 2010
>>
>> Efforts are underway to develop a low cost DMTD apparatus with
>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>> provide an appropriate measure of stability in this time range, or is
>> the ADEV estimate a more correct answer?
>>
>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>> from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
>> other counters behave similarly.
>>
>> I would appreciate any comments or observations on this topic.
>> My motivation is to discover the simplest scheme for making
>> stability measurements at this performance level; this is NOT
>> even close to the state-of-the-art, but can still be useful.
>>
>> Pete Rawson
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
>
>
>
BG
Bruce Griffiths
Sat, Feb 6, 2010 9:16 PM
The tight PLL method doesn't directly produce the average frequency over
Tau.
As explained in (see snapshot of relevant section):
NIST special Publication 1065 Handbook of Frequency Stability Analysis
http://tf.nist.gov/timefreq/general/pdf/2220.pdf
the average frequency deviations for averaging time Tau are needed for
the calculation.
You need to sample at a sufficiently high rate to avoid aliasing and
average (ie integrate) the individual EFC samples.
If one uses phase measures then the fluctuations in the frequency
averages can easily and directly calculated from the difference between
the phase measured at time intervals separated by Tau.
Bruce
WarrenS wrote:
Thus NIST and others quietly dropped this method several decades ago.
Could it be another reason?
I'll bet that was after they wanted to do better than 1e14 resolution
AND had unlimited amounts of time and Money,
Something most time Nuts are not blessed with. I Never said it was
the BEST way.
JUST given the goal, which was 1e13 in one second, there is not a
simpler and cheaper way to do it.
And nothing you said counter that point.
The frequency measures need to be integrated (either implicitly or
explicitly) to produce phase measures which can then be used to
calculate ADEV, MDEV etc.
Well ONE of us certainly has something backward.
To calculate ADEV, MDEV etc. YOU need Freq Differences.
The first thing that happens when phase is used is that it is turned
into Freq by taking the difference between each sample.
Integrated Freq data, which is what "Tight Phase-Lock Loop Method"
gives you directly (no Phase conversion needed),
Need not FIRST turned into Phase so that it can then be turned back
into Freq.
BUT in any case there is no difference in the noise, for a given
bandwidth, If you don't run out of digits and You have enough resolution.
The "Tight Phase-Lock Loop Method" can EASY get sub pS resolution,
which is better than most other ways.
AND don't need filters and slue rate control and multistage limiters
and on & on to do it, an RC works fine to replace all the stuff.
ws
----- Original Message ----- From: "Bruce Griffiths"
bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Saturday, February 06, 2010 12:11 PM
Subject: Re: [time-nuts] ADEV vs MDEV
Sounds good but you still haven't found its Achilles heel:
The frequency measures need to be integrated (either implicitly or
explicitly) to produce phase measures which can then be used to
calculate ADEV, MDEV etc.
The major problem is that integration amplifies the small errors that
are inevitably present.
In practice (except for very noisy sources) the technique isnt
particularly useful for Tau more than a few times the inverse PLL
bandwidth.
Thus NIST and others quietly dropped this method several decades ago.
This is alluded to in Steins recent paper availble on the Symmetricom
website:
The Allan Variance – Challenges and Opportunities
Bruce
WarrenS wrote:
I would appreciate any comments or observations on the topic of
apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making
stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the
Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that
level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter& Printer
with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high
end test equipment needed.
If you want performance that exceeds the best of most DMTD at low
Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset
voltage.
I must add this is not a popular solution (Or a general Purpose one)
but
IF you know analog and have a GOOD osc with EFC to use for the
reference,
as far as I've been able to determine it is the BEST SIMPLE answer
that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1
sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per
Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD
Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out
there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
The tight PLL method doesn't directly produce the average frequency over
Tau.
As explained in (see snapshot of relevant section):
NIST special Publication 1065 Handbook of Frequency Stability Analysis
<http://tf.nist.gov/timefreq/general/pdf/2220.pdf>
the average frequency deviations for averaging time Tau are needed for
the calculation.
You need to sample at a sufficiently high rate to avoid aliasing and
average (ie integrate) the individual EFC samples.
If one uses phase measures then the fluctuations in the frequency
averages can easily and directly calculated from the difference between
the phase measured at time intervals separated by Tau.
Bruce
WarrenS wrote:
> Bruce said:
>
>> Thus NIST and others quietly dropped this method several decades ago.
> Could it be another reason?
> I'll bet that was after they wanted to do better than 1e14 resolution
> AND had unlimited amounts of time and Money,
> Something most time Nuts are not blessed with. I Never said it was
> the BEST way.
> JUST given the goal, which was 1e13 in one second, there is not a
> simpler and cheaper way to do it.
> And nothing you said counter that point.
>
>
>> The frequency measures need to be integrated (either implicitly or
>> explicitly) to produce phase measures which can then be used to
>> calculate ADEV, MDEV etc.
>
> Well ONE of us certainly has something backward.
> To calculate ADEV, MDEV etc. YOU need Freq Differences.
> The first thing that happens when phase is used is that it is turned
> into Freq by taking the difference between each sample.
> Integrated Freq data, which is what "Tight Phase-Lock Loop Method"
> gives you directly (no Phase conversion needed),
> Need not FIRST turned into Phase so that it can then be turned back
> into Freq.
> BUT in any case there is no difference in the noise, for a given
> bandwidth, If you don't run out of digits and You have enough resolution.
> The "Tight Phase-Lock Loop Method" can EASY get sub pS resolution,
> which is better than most other ways.
> AND don't need filters and slue rate control and multistage limiters
> and on & on to do it, an RC works fine to replace all the stuff.
>
> ws
>
> *****************
>
> ----- Original Message ----- From: "Bruce Griffiths"
> <bruce.griffiths@xtra.co.nz>
> To: "Discussion of precise time and frequency measurement"
> <time-nuts@febo.com>
> Sent: Saturday, February 06, 2010 12:11 PM
> Subject: Re: [time-nuts] ADEV vs MDEV
>
>
>> Sounds good but you still haven't found its Achilles heel:
>>
>> The frequency measures need to be integrated (either implicitly or
>> explicitly) to produce phase measures which can then be used to
>> calculate ADEV, MDEV etc.
>> The major problem is that integration amplifies the small errors that
>> are inevitably present.
>> In practice (except for very noisy sources) the technique isnt
>> particularly useful for Tau more than a few times the inverse PLL
>> bandwidth.
>>
>> Thus NIST and others quietly dropped this method several decades ago.
>> This is alluded to in Steins recent paper availble on the Symmetricom
>> website:
>>
>> *The Allan Variance – Challenges and Opportunities*
>>
>>
>> Bruce
>>
>> WarrenS wrote:
>>> Peat said:
>>>
>>>> I would appreciate any comments or observations on the topic of
>>>> apparatus with demonstrated stability measurements.
>>>> My motivation is to discover the SIMPLEST scheme for making
>>>> stability measurements at the 1E-13 in 1s performance level.
>>>>
>>>
>>> If you accept that the measurement is going to limited by the
>>> Reference Osc,
>>> for Low COST and SIMPLE, with the ability to measure ADEVs at that
>>> level,
>>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
>>> Method of measuring Freq stability".
>>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>>
>>>
>>> By replacing the "Voltage to freq converter, Freq counter& Printer
>>> with a Radio shack type PC data logging DVM,
>>> It can be up and running from scratch in under an Hr, with no high
>>> end test equipment needed.
>>> If you want performance that exceeds the best of most DMTD at low
>>> Tau it takes a little more work
>>> and a higher speed oversampling ADC data logger and a good offset
>>> voltage.
>>>
>>> I must add this is not a popular solution (Or a general Purpose one)
>>> but
>>> IF you know analog and have a GOOD osc with EFC to use for the
>>> reference,
>>> as far as I've been able to determine it is the BEST SIMPLE answer
>>> that allows High performance.
>>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1
>>> sec (at 30 Hz Bandwidth)
>>>
>>> Basic modified NIST Block Diag attached:
>>> The NIST paper sums it up quite nicely:
>>> 'It is not difficult to achieve a sensitivity of a part in e14 per
>>> Hz resolution
>>> so one has excellent precision capabilities with this system.'
>>>
>>> This does not address your other question of ADEV vs MDEV,
>>> What I've described is just a simple way to get the Low cost, GOOD
>>> Raw data.
>>> What you then do with that Data is a different subject.
>>>
>>> You can run the raw data thru one of the many ADEV programs out
>>> there, 'Plotter' being my choice.
>>>
>>>
>>> Have fun
>>> ws
>>>
>>> *************
>>>
>>> [time-nuts] ADEV vs MDEV
>>> Pete Rawson peterawson at earthlink.net
>>> Sat Feb 6 03:59:18 UTC 2010
>>>
>>> Efforts are underway to develop a low cost DMTD apparatus with
>>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>>> provide an appropriate measure of stability in this time range, or is
>>> the ADEV estimate a more correct answer?
>>>
>>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>>> from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
>>> other counters behave similarly.
>>>
>>> I would appreciate any comments or observations on this topic.
>>> My motivation is to discover the simplest scheme for making
>>> stability measurements at this performance level; this is NOT
>>> even close to the state-of-the-art, but can still be useful.
>>>
>>> Pete Rawson
>>>
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>
>>
>>
>>
>>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
W
WarrenS
Sat, Feb 6, 2010 9:24 PM
"It's possible / likely for injection lock ... to be a problem ..."
Something I certainly worried about and tested for.
What I found (for MY case) is that injection lock is NOT a problem.
The reason being is that unlike most other ways, where the two OSC have to
be completely independent,
The tight loop approach forces the Two Osc to "Lock with something like 60 +
db gain,
so a little stray -80db injection lock coupling that would very much limit
other systems has
no measurable effect at e-13. Just one of the neat little side effects that
make the tight loop approach so simple.
"then a part in 10^14 is going to be at the 100 of nanovolts level."
For that example, just need to put a simple discrete 100 to 1 resistor
divider
in-between the control voltage and the EFC and now you have a nice workable
10uv.
BUT the bigger point is, probable not needed, cause you are NOT going to do
any better than the stability of the OSC with a grounded shorted EFC input.
as you said and I agree is so true:
"There is no perfect way to do any of this, only a lot of compromises ...
you need to watch out for".
But you did not offer any easier way to do it, which is what the original
request was for and my answer addressed.
This is the cheapest easiest way BY FAR to get high performance, at low tau,
ADEV numbers that I've seen.
ws
----- Original Message -----
From: "Bob Camp" lists@cq.nu
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Saturday, February 06, 2010 12:09 PM
Subject: Re: [time-nuts] ADEV vs MDEV
Hi
It's possible / likely to injection lock with the tight loop approach and
get data that's much better than reality. A lot depends on the specific
oscillators under test and the buffers (if any) between the oscillators
and mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14
is going to be at the 100 of nanovolts level. Certainly not impossible,
but it does present it's own set of issues. Lab gear to do it is
available, but not all that common. DC offsets and their temperature
coefficients along with thermocouple effects could make things exciting.
There is no perfect way to do any of this, only a lot of compromises here
or there. Each approach has stuff you need to watch out for.
Bob
From: "WarrenS" warrensjmail-one@yahoo.com
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Subject: Re: [time-nuts] ADEV vs MDEV
I would appreciate any comments or observations on the topic of apparatus
with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability
measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference
Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer with
a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end
test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it
takes a little more work
and a higher speed oversampling ADC data logger and a good offset
voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the
reference,
as far as I've been able to determine it is the BEST SIMPLE answer that
allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec
(at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz
resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw
data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there,
'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
> "It's possible / likely for injection lock ... to be a problem ..."
Something I certainly worried about and tested for.
What I found (for MY case) is that injection lock is NOT a problem.
The reason being is that unlike most other ways, where the two OSC have to
be completely independent,
The tight loop approach forces the Two Osc to "Lock with something like 60 +
db gain,
so a little stray -80db injection lock coupling that would very much limit
other systems has
no measurable effect at e-13. Just one of the neat little side effects that
make the tight loop approach so simple.
> "then a part in 10^14 is going to be at the 100 of nanovolts level."
For that example, just need to put a simple discrete 100 to 1 resistor
divider
in-between the control voltage and the EFC and now you have a nice workable
10uv.
BUT the bigger point is, probable not needed, cause you are NOT going to do
any better than the stability of the OSC with a grounded shorted EFC input.
as you said and I agree is so true:
> "There is no perfect way to do any of this, only a lot of compromises ...
> you need to watch out for".
But you did not offer any easier way to do it, which is what the original
request was for and my answer addressed.
This is the cheapest easiest way BY FAR to get high performance, at low tau,
ADEV numbers that I've seen.
ws
***************
----- Original Message -----
From: "Bob Camp" <lists@cq.nu>
To: "Discussion of precise time and frequency measurement"
<time-nuts@febo.com>
Sent: Saturday, February 06, 2010 12:09 PM
Subject: Re: [time-nuts] ADEV vs MDEV
> Hi
>
> It's possible / likely to injection lock with the tight loop approach and
> get data that's much better than reality. A lot depends on the specific
> oscillators under test and the buffers (if any) between the oscillators
> and mixer.
>
> If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14
> is going to be at the 100 of nanovolts level. Certainly not impossible,
> but it does present it's own set of issues. Lab gear to do it is
> available, but not all that common. DC offsets and their temperature
> coefficients along with thermocouple effects could make things exciting.
>
> There is no perfect way to do any of this, only a lot of compromises here
> or there. Each approach has stuff you need to watch out for.
>
> Bob
>
> --------------------------------------------------
> From: "WarrenS" <warrensjmail-one@yahoo.com>
> Sent: Saturday, February 06, 2010 2:19 PM
> To: "Discussion of precise time and frequency measurement"
> <time-nuts@febo.com>
> Subject: Re: [time-nuts] ADEV vs MDEV
>
>>
>> Peat said:
>>>I would appreciate any comments or observations on the topic of apparatus
>>>with demonstrated stability measurements.
>>>My motivation is to discover the SIMPLEST scheme for making stability
>>>measurements at the 1E-13 in 1s performance level.
>>
>>
>> If you accept that the measurement is going to limited by the Reference
>> Osc,
>> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop
>> Method of measuring Freq stability".
>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>
>>
>> By replacing the "Voltage to freq converter, Freq counter & Printer with
>> a Radio shack type PC data logging DVM,
>> It can be up and running from scratch in under an Hr, with no high end
>> test equipment needed.
>> If you want performance that exceeds the best of most DMTD at low Tau it
>> takes a little more work
>> and a higher speed oversampling ADC data logger and a good offset
>> voltage.
>>
>> I must add this is not a popular solution (Or a general Purpose one) but
>> IF you know analog and have a GOOD osc with EFC to use for the
>> reference,
>> as far as I've been able to determine it is the BEST SIMPLE answer that
>> allows High performance.
>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec
>> (at 30 Hz Bandwidth)
>>
>> Basic modified NIST Block Diag attached:
>> The NIST paper sums it up quite nicely:
>> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz
>> resolution
>> so one has excellent precision capabilities with this system.'
>>
>> This does not address your other question of ADEV vs MDEV,
>> What I've described is just a simple way to get the Low cost, GOOD Raw
>> data.
>> What you then do with that Data is a different subject.
>>
>> You can run the raw data thru one of the many ADEV programs out there,
>> 'Plotter' being my choice.
>>
>>
>> Have fun
>> ws
>>
>> *************
>>
>> [time-nuts] ADEV vs MDEV
>> Pete Rawson peterawson at earthlink.net
>> Sat Feb 6 03:59:18 UTC 2010
>>
>> Efforts are underway to develop a low cost DMTD apparatus with
>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>> provide an appropriate measure of stability in this time range, or is
>> the ADEV estimate a more correct answer?
>>
>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>> from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
>> other counters behave similarly.
>>
>> I would appreciate any comments or observations on this topic.
>> My motivation is to discover the simplest scheme for making
>> stability measurements at this performance level; this is NOT
>> even close to the state-of-the-art, but can still be useful.
>>
>> Pete Rawson
>>
BC
Bob Camp
Sat, Feb 6, 2010 10:03 PM
Hi
A straight heterodyne system will get you to the floor of most 10811's with a very simple (2 stage) limiter. As with the DMTD, the counter requirements aren't really all that severe.
Bob
On Feb 6, 2010, at 4:24 PM, WarrenS wrote:
"It's possible / likely for injection lock ... to be a problem ..."
Something I certainly worried about and tested for.
What I found (for MY case) is that injection lock is NOT a problem.
The reason being is that unlike most other ways, where the two OSC have to be completely independent,
The tight loop approach forces the Two Osc to "Lock with something like 60 + db gain,
so a little stray -80db injection lock coupling that would very much limit other systems has
no measurable effect at e-13. Just one of the neat little side effects that make the tight loop approach so simple.
"then a part in 10^14 is going to be at the 100 of nanovolts level."
For that example, just need to put a simple discrete 100 to 1 resistor divider
in-between the control voltage and the EFC and now you have a nice workable 10uv.
BUT the bigger point is, probable not needed, cause you are NOT going to do any better than the stability of the OSC with a grounded shorted EFC input.
as you said and I agree is so true:
"There is no perfect way to do any of this, only a lot of compromises ... you need to watch out for".
But you did not offer any easier way to do it, which is what the original request was for and my answer addressed.
This is the cheapest easiest way BY FAR to get high performance, at low tau, ADEV numbers that I've seen.
ws
----- Original Message ----- From: "Bob Camp" lists@cq.nu
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Saturday, February 06, 2010 12:09 PM
Subject: Re: [time-nuts] ADEV vs MDEV
Hi
It's possible / likely to injection lock with the tight loop approach and get data that's much better than reality. A lot depends on the specific oscillators under test and the buffers (if any) between the oscillators and mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is going to be at the 100 of nanovolts level. Certainly not impossible, but it does present it's own set of issues. Lab gear to do it is available, but not all that common. DC offsets and their temperature coefficients along with thermocouple effects could make things exciting.
There is no perfect way to do any of this, only a lot of compromises here or there. Each approach has stuff you need to watch out for.
Bob
From: "WarrenS" warrensjmail-one@yahoo.com
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Subject: Re: [time-nuts] ADEV vs MDEV
I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter & Printer with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
Hi
A straight heterodyne system will get you to the floor of most 10811's with a very simple (2 stage) limiter. As with the DMTD, the counter requirements aren't really all that severe.
Bob
On Feb 6, 2010, at 4:24 PM, WarrenS wrote:
>
>> "It's possible / likely for injection lock ... to be a problem ..."
> Something I certainly worried about and tested for.
> What I found (for MY case) is that injection lock is NOT a problem.
> The reason being is that unlike most other ways, where the two OSC have to be completely independent,
> The tight loop approach forces the Two Osc to "Lock with something like 60 + db gain,
> so a little stray -80db injection lock coupling that would very much limit other systems has
> no measurable effect at e-13. Just one of the neat little side effects that make the tight loop approach so simple.
>
>> "then a part in 10^14 is going to be at the 100 of nanovolts level."
> For that example, just need to put a simple discrete 100 to 1 resistor divider
> in-between the control voltage and the EFC and now you have a nice workable 10uv.
> BUT the bigger point is, probable not needed, cause you are NOT going to do any better than the stability of the OSC with a grounded shorted EFC input.
>
> as you said and I agree is so true:
>> "There is no perfect way to do any of this, only a lot of compromises ... you need to watch out for".
> But you did not offer any easier way to do it, which is what the original request was for and my answer addressed.
> This is the cheapest easiest way BY FAR to get high performance, at low tau, ADEV numbers that I've seen.
>
> ws
> ***************
>
> ----- Original Message ----- From: "Bob Camp" <lists@cq.nu>
> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
> Sent: Saturday, February 06, 2010 12:09 PM
> Subject: Re: [time-nuts] ADEV vs MDEV
>
>
>> Hi
>>
>> It's possible / likely to injection lock with the tight loop approach and get data that's much better than reality. A lot depends on the specific oscillators under test and the buffers (if any) between the oscillators and mixer.
>>
>> If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is going to be at the 100 of nanovolts level. Certainly not impossible, but it does present it's own set of issues. Lab gear to do it is available, but not all that common. DC offsets and their temperature coefficients along with thermocouple effects could make things exciting.
>>
>> There is no perfect way to do any of this, only a lot of compromises here or there. Each approach has stuff you need to watch out for.
>>
>> Bob
>>
>> --------------------------------------------------
>> From: "WarrenS" <warrensjmail-one@yahoo.com>
>> Sent: Saturday, February 06, 2010 2:19 PM
>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>> Subject: Re: [time-nuts] ADEV vs MDEV
>>
>>>
>>> Peat said:
>>>> I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
>>>> My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
>>>
>>>
>>> If you accept that the measurement is going to limited by the Reference Osc,
>>> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
>>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
>>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>>
>>>
>>> By replacing the "Voltage to freq converter, Freq counter & Printer with a Radio shack type PC data logging DVM,
>>> It can be up and running from scratch in under an Hr, with no high end test equipment needed.
>>> If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
>>> and a higher speed oversampling ADC data logger and a good offset voltage.
>>>
>>> I must add this is not a popular solution (Or a general Purpose one) but
>>> IF you know analog and have a GOOD osc with EFC to use for the reference,
>>> as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
>>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
>>>
>>> Basic modified NIST Block Diag attached:
>>> The NIST paper sums it up quite nicely:
>>> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
>>> so one has excellent precision capabilities with this system.'
>>>
>>> This does not address your other question of ADEV vs MDEV,
>>> What I've described is just a simple way to get the Low cost, GOOD Raw data.
>>> What you then do with that Data is a different subject.
>>>
>>> You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
>>>
>>>
>>> Have fun
>>> ws
>>>
>>> *************
>>>
>>> [time-nuts] ADEV vs MDEV
>>> Pete Rawson peterawson at earthlink.net
>>> Sat Feb 6 03:59:18 UTC 2010
>>>
>>> Efforts are underway to develop a low cost DMTD apparatus with
>>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>>> provide an appropriate measure of stability in this time range, or is
>>> the ADEV estimate a more correct answer?
>>>
>>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>>> from my CNT81showing MDEV < 1E-13 in 10s. and I believe the
>>> other counters behave similarly.
>>>
>>> I would appreciate any comments or observations on this topic.
>>> My motivation is to discover the simplest scheme for making
>>> stability measurements at this performance level; this is NOT
>>> even close to the state-of-the-art, but can still be useful.
>>>
>>> Pete Rawson
>>>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
BG
Bruce Griffiths
Sat, Feb 6, 2010 10:12 PM
The only major issue with DMTD systems is that they undersample the
phase fluctuations and hence are subject to aliasing effects.
The low pass filter has to have a bandwidth of the same order as the
beat frequency or the beat frequency signal will be significantly
attenuated.
Since the phase is only sampled once per beat frequency period the phase
fluctuations are undersampled.
Various attempts to use both zero crossings have not been successful.
In principle if one can overcome the increased phase shift tempco
associated with a bandpass filter, using a bandpass filter can in
principle ensure that the phase fluctuations are oversampled.
Bruce
Bob Camp wrote:
Hi
A straight heterodyne system will get you to the floor of most 10811's with a very simple (2 stage) limiter. As with the DMTD, the counter requirements aren't really all that severe.
Bob
On Feb 6, 2010, at 4:24 PM, WarrenS wrote:
"It's possible / likely for injection lock ... to be a problem ..."
Something I certainly worried about and tested for.
What I found (for MY case) is that injection lock is NOT a problem.
The reason being is that unlike most other ways, where the two OSC have to be completely independent,
The tight loop approach forces the Two Osc to "Lock with something like 60 + db gain,
so a little stray -80db injection lock coupling that would very much limit other systems has
no measurable effect at e-13. Just one of the neat little side effects that make the tight loop approach so simple.
"then a part in 10^14 is going to be at the 100 of nanovolts level."
For that example, just need to put a simple discrete 100 to 1 resistor divider
in-between the control voltage and the EFC and now you have a nice workable 10uv.
BUT the bigger point is, probable not needed, cause you are NOT going to do any better than the stability of the OSC with a grounded shorted EFC input.
as you said and I agree is so true:
"There is no perfect way to do any of this, only a lot of compromises ... you need to watch out for".
But you did not offer any easier way to do it, which is what the original request was for and my answer addressed.
This is the cheapest easiest way BY FAR to get high performance, at low tau, ADEV numbers that I've seen.
ws
----- Original Message ----- From: "Bob Camp"lists@cq.nu
To: "Discussion of precise time and frequency measurement"time-nuts@febo.com
Sent: Saturday, February 06, 2010 12:09 PM
Subject: Re: [time-nuts] ADEV vs MDEV
Hi
It's possible / likely to injection lock with the tight loop approach and get data that's much better than reality. A lot depends on the specific oscillators under test and the buffers (if any) between the oscillators and mixer.
If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is going to be at the 100 of nanovolts level. Certainly not impossible, but it does present it's own set of issues. Lab gear to do it is available, but not all that common. DC offsets and their temperature coefficients along with thermocouple effects could make things exciting.
There is no perfect way to do any of this, only a lot of compromises here or there. Each approach has stuff you need to watch out for.
Bob
From: "WarrenS"warrensjmail-one@yahoo.com
Sent: Saturday, February 06, 2010 2:19 PM
To: "Discussion of precise time and frequency measurement"time-nuts@febo.com
Subject: Re: [time-nuts] ADEV vs MDEV
I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
If you accept that the measurement is going to limited by the Reference Osc,
for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
By replacing the "Voltage to freq converter, Freq counter& Printer with a Radio shack type PC data logging DVM,
It can be up and running from scratch in under an Hr, with no high end test equipment needed.
If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
and a higher speed oversampling ADC data logger and a good offset voltage.
I must add this is not a popular solution (Or a general Purpose one) but
IF you know analog and have a GOOD osc with EFC to use for the reference,
as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
Basic modified NIST Block Diag attached:
The NIST paper sums it up quite nicely:
'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
so one has excellent precision capabilities with this system.'
This does not address your other question of ADEV vs MDEV,
What I've described is just a simple way to get the Low cost, GOOD Raw data.
What you then do with that Data is a different subject.
You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
Have fun
ws
[time-nuts] ADEV vs MDEV
Pete Rawson peterawson at earthlink.net
Sat Feb 6 03:59:18 UTC 2010
Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?
The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
other counters behave similarly.
I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.
Pete Rawson
The only major issue with DMTD systems is that they undersample the
phase fluctuations and hence are subject to aliasing effects.
The low pass filter has to have a bandwidth of the same order as the
beat frequency or the beat frequency signal will be significantly
attenuated.
Since the phase is only sampled once per beat frequency period the phase
fluctuations are undersampled.
Various attempts to use both zero crossings have not been successful.
In principle if one can overcome the increased phase shift tempco
associated with a bandpass filter, using a bandpass filter can in
principle ensure that the phase fluctuations are oversampled.
Bruce
Bob Camp wrote:
> Hi
>
> A straight heterodyne system will get you to the floor of most 10811's with a very simple (2 stage) limiter. As with the DMTD, the counter requirements aren't really all that severe.
>
> Bob
>
>
> On Feb 6, 2010, at 4:24 PM, WarrenS wrote:
>
>
>>
>>> "It's possible / likely for injection lock ... to be a problem ..."
>>>
>> Something I certainly worried about and tested for.
>> What I found (for MY case) is that injection lock is NOT a problem.
>> The reason being is that unlike most other ways, where the two OSC have to be completely independent,
>> The tight loop approach forces the Two Osc to "Lock with something like 60 + db gain,
>> so a little stray -80db injection lock coupling that would very much limit other systems has
>> no measurable effect at e-13. Just one of the neat little side effects that make the tight loop approach so simple.
>>
>>
>>> "then a part in 10^14 is going to be at the 100 of nanovolts level."
>>>
>> For that example, just need to put a simple discrete 100 to 1 resistor divider
>> in-between the control voltage and the EFC and now you have a nice workable 10uv.
>> BUT the bigger point is, probable not needed, cause you are NOT going to do any better than the stability of the OSC with a grounded shorted EFC input.
>>
>> as you said and I agree is so true:
>>
>>> "There is no perfect way to do any of this, only a lot of compromises ... you need to watch out for".
>>>
>> But you did not offer any easier way to do it, which is what the original request was for and my answer addressed.
>> This is the cheapest easiest way BY FAR to get high performance, at low tau, ADEV numbers that I've seen.
>>
>> ws
>> ***************
>>
>> ----- Original Message ----- From: "Bob Camp"<lists@cq.nu>
>> To: "Discussion of precise time and frequency measurement"<time-nuts@febo.com>
>> Sent: Saturday, February 06, 2010 12:09 PM
>> Subject: Re: [time-nuts] ADEV vs MDEV
>>
>>
>>
>>> Hi
>>>
>>> It's possible / likely to injection lock with the tight loop approach and get data that's much better than reality. A lot depends on the specific oscillators under test and the buffers (if any) between the oscillators and mixer.
>>>
>>> If your OCVCXO has a tuning slope of 0.1 ppm / volt then a part in 10^14 is going to be at the 100 of nanovolts level. Certainly not impossible, but it does present it's own set of issues. Lab gear to do it is available, but not all that common. DC offsets and their temperature coefficients along with thermocouple effects could make things exciting.
>>>
>>> There is no perfect way to do any of this, only a lot of compromises here or there. Each approach has stuff you need to watch out for.
>>>
>>> Bob
>>>
>>> --------------------------------------------------
>>> From: "WarrenS"<warrensjmail-one@yahoo.com>
>>> Sent: Saturday, February 06, 2010 2:19 PM
>>> To: "Discussion of precise time and frequency measurement"<time-nuts@febo.com>
>>> Subject: Re: [time-nuts] ADEV vs MDEV
>>>
>>>
>>>> Peat said:
>>>>
>>>>> I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
>>>>> My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s performance level.
>>>>>
>>>>
>>>> If you accept that the measurement is going to limited by the Reference Osc,
>>>> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
>>>> Can't beat a simple analog version of NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
>>>> http://tf.nist.gov/phase/Properties/one.htm#oneone Fig 1.7
>>>>
>>>>
>>>> By replacing the "Voltage to freq converter, Freq counter& Printer with a Radio shack type PC data logging DVM,
>>>> It can be up and running from scratch in under an Hr, with no high end test equipment needed.
>>>> If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
>>>> and a higher speed oversampling ADC data logger and a good offset voltage.
>>>>
>>>> I must add this is not a popular solution (Or a general Purpose one) but
>>>> IF you know analog and have a GOOD osc with EFC to use for the reference,
>>>> as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
>>>> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
>>>>
>>>> Basic modified NIST Block Diag attached:
>>>> The NIST paper sums it up quite nicely:
>>>> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
>>>> so one has excellent precision capabilities with this system.'
>>>>
>>>> This does not address your other question of ADEV vs MDEV,
>>>> What I've described is just a simple way to get the Low cost, GOOD Raw data.
>>>> What you then do with that Data is a different subject.
>>>>
>>>> You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
>>>>
>>>>
>>>> Have fun
>>>> ws
>>>>
>>>> *************
>>>>
>>>> [time-nuts] ADEV vs MDEV
>>>> Pete Rawson peterawson at earthlink.net
>>>> Sat Feb 6 03:59:18 UTC 2010
>>>>
>>>> Efforts are underway to develop a low cost DMTD apparatus with
>>>> demonstrated stability measurements of 1E-13 in 1s. It seems that
>>>> existing TI counters can reach this goal in 10s. (using MDEV estimate
>>>> or 100+s. using ADEV estimate). The question is; does the MDEV tool
>>>> provide an appropriate measure of stability in this time range, or is
>>>> the ADEV estimate a more correct answer?
>>>>
>>>> The TI performance I'm referring to is the 20-25 ps, single shot TI,
>>>> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
>>>> from my CNT81showing MDEV< 1E-13 in 10s. and I believe the
>>>> other counters behave similarly.
>>>>
>>>> I would appreciate any comments or observations on this topic.
>>>> My motivation is to discover the simplest scheme for making
>>>> stability measurements at this performance level; this is NOT
>>>> even close to the state-of-the-art, but can still be useful.
>>>>
>>>> Pete Rawson
>>>>
>>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>