RE
Randy Evans
Sat, Aug 23, 2014 2:42 AM
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit yet -
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data dumper
program to get the CAL data from my 3458A. Should be a busy weekend.
Randy
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit yet -
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data dumper
program to get the CAL data from my 3458A. Should be a busy weekend.
Randy
TM
Todd Micallef
Sat, Aug 23, 2014 3:06 AM
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler tool
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs regardless
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will typically
have some warranty if I remember correctly. Moving forward I will only use
2 12v batteries and pre-charge them on a battery charger to equalize them
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I had
a few that looked questionable. So far, I have replaced all the big caps on
the pre-regulator and regulator boards. My feeling is that once these go
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans randyevans2688@gmail.com
wrote:
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit yet -
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data dumper
program to get the CAL data from my 3458A. Should be a busy weekend.
Randy
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler tool
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs regardless
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will typically
have some warranty if I remember correctly. Moving forward I will only use
2 12v batteries and pre-charge them on a battery charger to equalize them
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I had
a few that looked questionable. So far, I have replaced all the big caps on
the pre-regulator and regulator boards. My feeling is that once these go
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <randyevans2688@gmail.com>
wrote:
> I received my Fluke 732A today. Just powered it up but it needs new
> batteries. Any suggestions for sources (I haven't opened up the unit yet -
> I want to make sure it works before doing that). Also received the
> ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data dumper
> program to get the CAL data from my 3458A. Should be a busy weekend.
>
> Randy
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
RE
Randy Evans
Sat, Aug 23, 2014 4:03 AM
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they are
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but they
slightly disagree. I am like the man with two watches that disagree on the
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com wrote:
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler tool
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs regardless
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will typically
have some warranty if I remember correctly. Moving forward I will only use
2 12v batteries and pre-charge them on a battery charger to equalize them
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I had
a few that looked questionable. So far, I have replaced all the big caps on
the pre-regulator and regulator boards. My feeling is that once these go
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans randyevans2688@gmail.com
wrote:
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they are
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but they
slightly disagree. I am like the man with two watches that disagree on the
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com> wrote:
> Randy,
>
> You have two possible choices. It can be configured with 4 x 6v 4Ah
> batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> modified the battery pack already. A couple of mine needed a nibbler tool
> to remove enough of the aluminum cover that fits over the tops of the
> batteries. The original cover will short out to the battery tabs regardless
> of the battery configuration if this is not done.
>
> You can find larger capacity batteries that will give you slightly more
> battery life. I lost a couple sets of mail-order batteries after a few
> extended outages. I would recommend going with locally bought batteries
> instead of the cheaper mail order. My local Batteries Plus will typically
> have some warranty if I remember correctly. Moving forward I will only use
> 2 12v batteries and pre-charge them on a battery charger to equalize them
> before putting them in the 732A. I think the cheap batteries did not
> discharge equally, and would not recover when power was applied.
>
> Inspect the back plane for damaged traces and look at the capacitors. I had
> a few that looked questionable. So far, I have replaced all the big caps on
> the pre-regulator and regulator boards. My feeling is that once these go
> online, they should run as long as possible between repairs.
>
> The battery charger circuit may need adjusting. I tweaked mine and it
> seemed to work fine.
>
> Todd
>
>
> On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > I received my Fluke 732A today. Just powered it up but it needs new
> > batteries. Any suggestions for sources (I haven't opened up the unit
> yet -
> > I want to make sure it works before doing that). Also received the
> > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
> dumper
> > program to get the CAL data from my 3458A. Should be a busy weekend.
> >
> > Randy
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
A
acbern@gmx.de
Sat, Aug 23, 2014 6:29 AM
Randy,
I have replaced batteries in two units so far, which I bought on ebay (6V) and they fitted absolutelly perfect, no need to rework an metal and there was o isue with shortages. So there certainly are standard 6V batteries that fit without mods. I always try to keep things in original config, if possible.
Also, you do not want to adjust the 732a trimmers. It only gets unstable, if you are unlucky. I discussed this with some cal labs, same answer. What you could do is to adjust the jumper settings for coarse adjustment, if need be.
Adrian
Gesendet: Samstag, 23. August 2014 um 04:03 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they are
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but they
slightly disagree. I am like the man with two watches that disagree on the
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com wrote:
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler tool
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs regardless
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will typically
have some warranty if I remember correctly. Moving forward I will only use
2 12v batteries and pre-charge them on a battery charger to equalize them
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I had
a few that looked questionable. So far, I have replaced all the big caps on
the pre-regulator and regulator boards. My feeling is that once these go
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans randyevans2688@gmail.com
wrote:
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
Randy,
I have replaced batteries in two units so far, which I bought on ebay (6V) and they fitted absolutelly perfect, no need to rework an metal and there was o isue with shortages. So there certainly are standard 6V batteries that fit without mods. I always try to keep things in original config, if possible.
Also, you do not want to adjust the 732a trimmers. It only gets unstable, if you are unlucky. I discussed this with some cal labs, same answer. What you could do is to adjust the jumper settings for coarse adjustment, if need be.
Adrian
> Gesendet: Samstag, 23. August 2014 um 04:03 Uhr
> Von: "Randy Evans" <randyevans2688@gmail.com>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> Todd,
>
> Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
> keep topped off and they have very low current draw (~2 to 3 mA at 13.5
> VDC) when charged and at their float voltage, so I am pretty sure they are
> in good condition. I will look at getting those in the units after I
> ascertain the condition of the 732.
>
> So now I have a what appears to be a functioning 3458A and a 732A but they
> slightly disagree. I am like the man with two watches that disagree on the
> time - which is correct? For the moment, i am only concerned with
> stability. The need for absolute accuracy will come later.
>
> Randy
>
>
> On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com> wrote:
>
> > Randy,
> >
> > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> > modified the battery pack already. A couple of mine needed a nibbler tool
> > to remove enough of the aluminum cover that fits over the tops of the
> > batteries. The original cover will short out to the battery tabs regardless
> > of the battery configuration if this is not done.
> >
> > You can find larger capacity batteries that will give you slightly more
> > battery life. I lost a couple sets of mail-order batteries after a few
> > extended outages. I would recommend going with locally bought batteries
> > instead of the cheaper mail order. My local Batteries Plus will typically
> > have some warranty if I remember correctly. Moving forward I will only use
> > 2 12v batteries and pre-charge them on a battery charger to equalize them
> > before putting them in the 732A. I think the cheap batteries did not
> > discharge equally, and would not recover when power was applied.
> >
> > Inspect the back plane for damaged traces and look at the capacitors. I had
> > a few that looked questionable. So far, I have replaced all the big caps on
> > the pre-regulator and regulator boards. My feeling is that once these go
> > online, they should run as long as possible between repairs.
> >
> > The battery charger circuit may need adjusting. I tweaked mine and it
> > seemed to work fine.
> >
> > Todd
> >
> >
> > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > I received my Fluke 732A today. Just powered it up but it needs new
> > > batteries. Any suggestions for sources (I haven't opened up the unit
> > yet -
> > > I want to make sure it works before doing that). Also received the
> > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
> > dumper
> > > program to get the CAL data from my 3458A. Should be a busy weekend.
> > >
> > > Randy
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
BG
Bill Gold
Sat, Aug 23, 2014 5:21 PM
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and others
that are in this size and package. Order from one of the usual electronics
distributors like Allied, Mouser, Digikey. This is a very common battery as
it is used in a lot of "EXIT" signs so they are lighted when the power goes
out. I don't see how the 12V 7AH will fit as they are too large. I guess
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
but the terminals are in the wrong place so you will have to "nibble" out
the aluminum plate that holds them in the 732A battery pack. You have to be
careful if you use the 12v 5AH as you will have 4 extra battery connection
leads to deal with and connect correctly. I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the "CAL" light goes
out when AC power is not applied. So shipping to Cal Lab can be a problem
if it is a distance away, or you have to use a shipper like UPS or FEDEX and
you ship the night before and then use their "Morning delivery" and the Cal
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the "ext
power" plug to last longer. I have seen it done. The issue is to get the
Cal Lab to charge the extra battery before they ship the 732A back to you.
When you remove and work on the battery pack always have the AC power
plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply (battery)
voltage drops below about 21 volts the "CAL" light will go out. Below that
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the Reference
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification. When
the semiconductor junctions are unbiased and cool off when power is lost,
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years that
these units have been powered up, this won't happen and when power is lost
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this. If
you don't have "low thermal" connection leads you can experience uV changes
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads. I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate). You have to allow at least a minute or more before being able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A. Using the MATH function and all of the data you
can collect. That was after waiting for several minutes after plugging in
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they are
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but they
slightly disagree. I am like the man with two watches that disagree on
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will only
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I
a few that looked questionable. So far, I have replaced all the big caps
the pre-regulator and regulator boards. My feeling is that once these go
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans randyevans2688@gmail.com
wrote:
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
and follow the instructions there.
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and others
that are in this size and package. Order from one of the usual electronics
distributors like Allied, Mouser, Digikey. This is a very common battery as
it is used in a lot of "EXIT" signs so they are lighted when the power goes
out. I don't see how the 12V 7AH will fit as they are too large. I guess
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
but the terminals are in the wrong place so you will have to "nibble" out
the aluminum plate that holds them in the 732A battery pack. You have to be
careful if you use the 12v 5AH as you will have 4 extra battery connection
leads to deal with and connect correctly. I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the "CAL" light goes
out when AC power is not applied. So shipping to Cal Lab can be a problem
if it is a distance away, or you have to use a shipper like UPS or FEDEX and
you ship the night before and then use their "Morning delivery" and the Cal
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the "ext
power" plug to last longer. I have seen it done. The issue is to get the
Cal Lab to charge the extra battery before they ship the 732A back to you.
When you remove and work on the battery pack always have the AC power
plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply (battery)
voltage drops below about 21 volts the "CAL" light will go out. Below that
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the Reference
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification. When
the semiconductor junctions are unbiased and cool off when power is lost,
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years that
these units have been powered up, this won't happen and when power is lost
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this. If
you don't have "low thermal" connection leads you can experience uV changes
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads. I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate). You have to allow at least a minute or more before being able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A. Using the MATH function and all of the data you
can collect. That was after waiting for several minutes after plugging in
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
> Todd,
>
> Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
> keep topped off and they have very low current draw (~2 to 3 mA at 13.5
> VDC) when charged and at their float voltage, so I am pretty sure they are
> in good condition. I will look at getting those in the units after I
> ascertain the condition of the 732.
>
> So now I have a what appears to be a functioning 3458A and a 732A but they
> slightly disagree. I am like the man with two watches that disagree on
the
> time - which is correct? For the moment, i am only concerned with
> stability. The need for absolute accuracy will come later.
>
> Randy
>
>
> On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
wrote:
>
> > Randy,
> >
> > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> > modified the battery pack already. A couple of mine needed a nibbler
tool
> > to remove enough of the aluminum cover that fits over the tops of the
> > batteries. The original cover will short out to the battery tabs
regardless
> > of the battery configuration if this is not done.
> >
> > You can find larger capacity batteries that will give you slightly more
> > battery life. I lost a couple sets of mail-order batteries after a few
> > extended outages. I would recommend going with locally bought batteries
> > instead of the cheaper mail order. My local Batteries Plus will
typically
> > have some warranty if I remember correctly. Moving forward I will only
use
> > 2 12v batteries and pre-charge them on a battery charger to equalize
them
> > before putting them in the 732A. I think the cheap batteries did not
> > discharge equally, and would not recover when power was applied.
> >
> > Inspect the back plane for damaged traces and look at the capacitors. I
had
> > a few that looked questionable. So far, I have replaced all the big caps
on
> > the pre-regulator and regulator boards. My feeling is that once these go
> > online, they should run as long as possible between repairs.
> >
> > The battery charger circuit may need adjusting. I tweaked mine and it
> > seemed to work fine.
> >
> > Todd
> >
> >
> > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > I received my Fluke 732A today. Just powered it up but it needs new
> > > batteries. Any suggestions for sources (I haven't opened up the unit
> > yet -
> > > I want to make sure it works before doing that). Also received the
> > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
> > dumper
> > > program to get the CAL data from my 3458A. Should be a busy weekend.
> > >
> > > Randy
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RE
Randy Evans
Sun, Aug 24, 2014 2:22 PM
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If I
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume this
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of 100
and 1000) is much greater than what you are measuring on your system. Not
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather large
differences (this is after an ACAL). I need to find some better cables to
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual electronics
distributors like Allied, Mouser, Digikey. This is a very common battery
as
it is used in a lot of "EXIT" signs so they are lighted when the power goes
out. I don't see how the 12V 7AH will fit as they are too large. I guess
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
but the terminals are in the wrong place so you will have to "nibble" out
the aluminum plate that holds them in the 732A battery pack. You have to
be
careful if you use the 12v 5AH as you will have 4 extra battery connection
leads to deal with and connect correctly. I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the "CAL" light goes
out when AC power is not applied. So shipping to Cal Lab can be a problem
if it is a distance away, or you have to use a shipper like UPS or FEDEX
and
you ship the night before and then use their "Morning delivery" and the Cal
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the "ext
power" plug to last longer. I have seen it done. The issue is to get the
Cal Lab to charge the extra battery before they ship the 732A back to you.
When you remove and work on the battery pack always have the AC power
plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply (battery)
voltage drops below about 21 volts the "CAL" light will go out. Below that
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the Reference
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification. When
the semiconductor junctions are unbiased and cool off when power is lost,
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is lost
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this. If
you don't have "low thermal" connection leads you can experience uV changes
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads. I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A. Using the MATH function and all of the data you
can collect. That was after waiting for several minutes after plugging in
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree on
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will only
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once these
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
and follow the instructions there.
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If I
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume this
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of 100
and 1000) is much greater than what you are measuring on your system. Not
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather large
differences (this is after an ACAL). I need to find some better cables to
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
> Randy:
>
> I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> others
> that are in this size and package. Order from one of the usual electronics
> distributors like Allied, Mouser, Digikey. This is a very common battery
> as
> it is used in a lot of "EXIT" signs so they are lighted when the power goes
> out. I don't see how the 12V 7AH will fit as they are too large. I guess
> you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
> but the terminals are in the wrong place so you will have to "nibble" out
> the aluminum plate that holds them in the 732A battery pack. You have to
> be
> careful if you use the 12v 5AH as you will have 4 extra battery connection
> leads to deal with and connect correctly. I would stick with the 6V 4AH.
> New batteries will last around 12 to 14 hours before the "CAL" light goes
> out when AC power is not applied. So shipping to Cal Lab can be a problem
> if it is a distance away, or you have to use a shipper like UPS or FEDEX
> and
> you ship the night before and then use their "Morning delivery" and the Cal
> Lab is expecting your 732A. Same on the way back to you. Of course you
> could always strap another battery on the 732A and hook it up to the "ext
> power" plug to last longer. I have seen it done. The issue is to get the
> Cal Lab to charge the extra battery before they ship the 732A back to you.
>
> When you remove and work on the battery pack always have the AC power
> plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
> is working.
>
> The "CAL" light is to indicate that power has not been lost to the
> Reference Amp or other associated circuits. When the raw supply (battery)
> voltage drops below about 21 volts the "CAL" light will go out. Below that
> voltage the heater circuits will not work correctly and the 18.6 volt
> regulated supply will not regulate. The requirement is that the Reference
> Amp be kept "alive" at all times to maintain the output voltage that was
> measured at the time of the most recent Calibration or Certification. When
> the semiconductor junctions are unbiased and cool off when power is lost,
> and then power is restored the result will be a different 10 volts than
> before the power failure. My experience is that after all of the years
> that
> these units have been powered up, this won't happen and when power is lost
> and then restored, even months later, the 732A will come back to almost
> exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
> after 24 hours of "warm up".
>
> What type of hookup leads are you using when measuring the 1 volt
> output? What is the PLC set to? I always use 100 PLC to measure this. If
> you don't have "low thermal" connection leads you can experience uV changes
> for a minute or more after plugging in the leads due to the "thermals"
> generated because of the difference in temperature between the banana jacks
> on the 732A and the banana plugs of the leads. I have found that even just
> plugging in the lead will generate a thermal difference because of
> difference of temps and some heating due to the physical act of just
> inserting the plug because of friction between the jack and plug (my theory
> at any rate). You have to allow at least a minute or more before being
> able
> to make a measurement after plugging in the leads. I just measured the
> variation of the 1 volt output of my 732A and using my 3458A and I got a
> total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
> volt range of the 3458A. Using the MATH function and all of the data you
> can collect. That was after waiting for several minutes after plugging in
> the leads.
>
> I hope all of this helps.
>
> Bill
>
>
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Friday, August 22, 2014 9:03 PM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Todd,
> >
> > Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
> > keep topped off and they have very low current draw (~2 to 3 mA at 13.5
> > VDC) when charged and at their float voltage, so I am pretty sure they
> are
> > in good condition. I will look at getting those in the units after I
> > ascertain the condition of the 732.
> >
> > So now I have a what appears to be a functioning 3458A and a 732A but
> they
> > slightly disagree. I am like the man with two watches that disagree on
> the
> > time - which is correct? For the moment, i am only concerned with
> > stability. The need for absolute accuracy will come later.
> >
> > Randy
> >
> >
> > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> wrote:
> >
> > > Randy,
> > >
> > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> > > modified the battery pack already. A couple of mine needed a nibbler
> tool
> > > to remove enough of the aluminum cover that fits over the tops of the
> > > batteries. The original cover will short out to the battery tabs
> regardless
> > > of the battery configuration if this is not done.
> > >
> > > You can find larger capacity batteries that will give you slightly more
> > > battery life. I lost a couple sets of mail-order batteries after a few
> > > extended outages. I would recommend going with locally bought batteries
> > > instead of the cheaper mail order. My local Batteries Plus will
> typically
> > > have some warranty if I remember correctly. Moving forward I will only
> use
> > > 2 12v batteries and pre-charge them on a battery charger to equalize
> them
> > > before putting them in the 732A. I think the cheap batteries did not
> > > discharge equally, and would not recover when power was applied.
> > >
> > > Inspect the back plane for damaged traces and look at the capacitors. I
> had
> > > a few that looked questionable. So far, I have replaced all the big
> caps
> on
> > > the pre-regulator and regulator boards. My feeling is that once these
> go
> > > online, they should run as long as possible between repairs.
> > >
> > > The battery charger circuit may need adjusting. I tweaked mine and it
> > > seemed to work fine.
> > >
> > > Todd
> > >
> > >
> > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> randyevans2688@gmail.com>
> > > wrote:
> > >
> > > > I received my Fluke 732A today. Just powered it up but it needs new
> > > > batteries. Any suggestions for sources (I haven't opened up the unit
> > > yet -
> > > > I want to make sure it works before doing that). Also received the
> > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
> > > dumper
> > > > program to get the CAL data from my 3458A. Should be a busy weekend.
> > > >
> > > > Randy
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
TM
Todd Micallef
Sun, Aug 24, 2014 3:13 PM
Bill,
I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes,
I was mistaken. The 12V 5ah batteries are the ones that I am using in one
of my 732A. There is more play inside the tray with the 12v batteries by
several mm as compared to the 6v which only has a 2-3mm. The 12v conversion
is not difficult, but it is easier if the battery tray has already been
machined for the 12v batteries. The battery tabs don't line up well with
the existing holes and need to be widened. Two additional holes must also
be added.
Sourcing the batteries locally is more of a convenience than waiting for
the delivery. The el-cheapo 6v batteries are a waste of money and I have 8
batteries so far to prove it. They died the first time I had a 6hr outage
and would not hold a charge after that.
Todd
On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual electronics
distributors like Allied, Mouser, Digikey. This is a very common battery
as
it is used in a lot of "EXIT" signs so they are lighted when the power goes
out. I don't see how the 12V 7AH will fit as they are too large. I guess
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
but the terminals are in the wrong place so you will have to "nibble" out
the aluminum plate that holds them in the 732A battery pack. You have to
be
careful if you use the 12v 5AH as you will have 4 extra battery connection
leads to deal with and connect correctly. I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the "CAL" light goes
out when AC power is not applied. So shipping to Cal Lab can be a problem
if it is a distance away, or you have to use a shipper like UPS or FEDEX
and
you ship the night before and then use their "Morning delivery" and the Cal
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the "ext
power" plug to last longer. I have seen it done. The issue is to get the
Cal Lab to charge the extra battery before they ship the 732A back to you.
When you remove and work on the battery pack always have the AC power
plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply (battery)
voltage drops below about 21 volts the "CAL" light will go out. Below that
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the Reference
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification. When
the semiconductor junctions are unbiased and cool off when power is lost,
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is lost
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this. If
you don't have "low thermal" connection leads you can experience uV changes
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads. I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A. Using the MATH function and all of the data you
can collect. That was after waiting for several minutes after plugging in
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree on
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will only
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once these
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
and follow the instructions there.
Bill,
I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes,
I was mistaken. The 12V 5ah batteries are the ones that I am using in one
of my 732A. There is more play inside the tray with the 12v batteries by
several mm as compared to the 6v which only has a 2-3mm. The 12v conversion
is not difficult, but it is easier if the battery tray has already been
machined for the 12v batteries. The battery tabs don't line up well with
the existing holes and need to be widened. Two additional holes must also
be added.
Sourcing the batteries locally is more of a convenience than waiting for
the delivery. The el-cheapo 6v batteries are a waste of money and I have 8
batteries so far to prove it. They died the first time I had a 6hr outage
and would not hold a charge after that.
Todd
On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold <wpgold3637@att.net> wrote:
> Randy:
>
> I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> others
> that are in this size and package. Order from one of the usual electronics
> distributors like Allied, Mouser, Digikey. This is a very common battery
> as
> it is used in a lot of "EXIT" signs so they are lighted when the power goes
> out. I don't see how the 12V 7AH will fit as they are too large. I guess
> you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
> but the terminals are in the wrong place so you will have to "nibble" out
> the aluminum plate that holds them in the 732A battery pack. You have to
> be
> careful if you use the 12v 5AH as you will have 4 extra battery connection
> leads to deal with and connect correctly. I would stick with the 6V 4AH.
> New batteries will last around 12 to 14 hours before the "CAL" light goes
> out when AC power is not applied. So shipping to Cal Lab can be a problem
> if it is a distance away, or you have to use a shipper like UPS or FEDEX
> and
> you ship the night before and then use their "Morning delivery" and the Cal
> Lab is expecting your 732A. Same on the way back to you. Of course you
> could always strap another battery on the 732A and hook it up to the "ext
> power" plug to last longer. I have seen it done. The issue is to get the
> Cal Lab to charge the extra battery before they ship the 732A back to you.
>
> When you remove and work on the battery pack always have the AC power
> plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
> is working.
>
> The "CAL" light is to indicate that power has not been lost to the
> Reference Amp or other associated circuits. When the raw supply (battery)
> voltage drops below about 21 volts the "CAL" light will go out. Below that
> voltage the heater circuits will not work correctly and the 18.6 volt
> regulated supply will not regulate. The requirement is that the Reference
> Amp be kept "alive" at all times to maintain the output voltage that was
> measured at the time of the most recent Calibration or Certification. When
> the semiconductor junctions are unbiased and cool off when power is lost,
> and then power is restored the result will be a different 10 volts than
> before the power failure. My experience is that after all of the years
> that
> these units have been powered up, this won't happen and when power is lost
> and then restored, even months later, the 732A will come back to almost
> exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
> after 24 hours of "warm up".
>
> What type of hookup leads are you using when measuring the 1 volt
> output? What is the PLC set to? I always use 100 PLC to measure this. If
> you don't have "low thermal" connection leads you can experience uV changes
> for a minute or more after plugging in the leads due to the "thermals"
> generated because of the difference in temperature between the banana jacks
> on the 732A and the banana plugs of the leads. I have found that even just
> plugging in the lead will generate a thermal difference because of
> difference of temps and some heating due to the physical act of just
> inserting the plug because of friction between the jack and plug (my theory
> at any rate). You have to allow at least a minute or more before being
> able
> to make a measurement after plugging in the leads. I just measured the
> variation of the 1 volt output of my 732A and using my 3458A and I got a
> total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
> volt range of the 3458A. Using the MATH function and all of the data you
> can collect. That was after waiting for several minutes after plugging in
> the leads.
>
> I hope all of this helps.
>
> Bill
>
>
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Friday, August 22, 2014 9:03 PM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Todd,
> >
> > Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
> > keep topped off and they have very low current draw (~2 to 3 mA at 13.5
> > VDC) when charged and at their float voltage, so I am pretty sure they
> are
> > in good condition. I will look at getting those in the units after I
> > ascertain the condition of the 732.
> >
> > So now I have a what appears to be a functioning 3458A and a 732A but
> they
> > slightly disagree. I am like the man with two watches that disagree on
> the
> > time - which is correct? For the moment, i am only concerned with
> > stability. The need for absolute accuracy will come later.
> >
> > Randy
> >
> >
> > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> wrote:
> >
> > > Randy,
> > >
> > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> > > modified the battery pack already. A couple of mine needed a nibbler
> tool
> > > to remove enough of the aluminum cover that fits over the tops of the
> > > batteries. The original cover will short out to the battery tabs
> regardless
> > > of the battery configuration if this is not done.
> > >
> > > You can find larger capacity batteries that will give you slightly more
> > > battery life. I lost a couple sets of mail-order batteries after a few
> > > extended outages. I would recommend going with locally bought batteries
> > > instead of the cheaper mail order. My local Batteries Plus will
> typically
> > > have some warranty if I remember correctly. Moving forward I will only
> use
> > > 2 12v batteries and pre-charge them on a battery charger to equalize
> them
> > > before putting them in the 732A. I think the cheap batteries did not
> > > discharge equally, and would not recover when power was applied.
> > >
> > > Inspect the back plane for damaged traces and look at the capacitors. I
> had
> > > a few that looked questionable. So far, I have replaced all the big
> caps
> on
> > > the pre-regulator and regulator boards. My feeling is that once these
> go
> > > online, they should run as long as possible between repairs.
> > >
> > > The battery charger circuit may need adjusting. I tweaked mine and it
> > > seemed to work fine.
> > >
> > > Todd
> > >
> > >
> > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> randyevans2688@gmail.com>
> > > wrote:
> > >
> > > > I received my Fluke 732A today. Just powered it up but it needs new
> > > > batteries. Any suggestions for sources (I haven't opened up the unit
> > > yet -
> > > > I want to make sure it works before doing that). Also received the
> > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
> > > dumper
> > > > program to get the CAL data from my 3458A. Should be a busy weekend.
> > > >
> > > > Randy
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
CB
Charles Black
Sun, Aug 24, 2014 7:04 PM
Bill,
I have never been able to use banana plugs of any kind, with my 3458A,
due to their excessive thermal mass. I use 24 gauge wires to make a
twisted pair by hand with no connectors for precision voltage work. As
brought out in an earlier post by acbern, it doesn't matter about
plating or solder at least as long as your twisted pairs are made from
the same wire so the thermocouples balance out.
I am a big fan of M27500 shielded twisted pair wire for everyday use. It
is durable, heat resistant and capable of very low voltage use. The
insulator is Tefzel and the plating used to be silver but may be
something else now. It take a few minutes work for proper termination
but it can be done for free. I tend to use a lot of the 24 gauge M27500
STP wire since it is fast to come to temperature, flexible and durable
enough to last quite awhile.
eBay always has quite a few small lengths available and Sierra Nevada
Products has a bunch of it. Sierra Nevada Products says that it is
Teflon but I think it may be Tefzel. I can help with termination
instructions if wanted.
Charlie
On 8/24/2014 7:22 AM, Randy Evans wrote:
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If I
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume this
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of 100
and 1000) is much greater than what you are measuring on your system. Not
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather large
differences (this is after an ACAL). I need to find some better cables to
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual electronics
distributors like Allied, Mouser, Digikey. This is a very common battery
as
it is used in a lot of "EXIT" signs so they are lighted when the power goes
out. I don't see how the 12V 7AH will fit as they are too large. I guess
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
but the terminals are in the wrong place so you will have to "nibble" out
the aluminum plate that holds them in the 732A battery pack. You have to
be
careful if you use the 12v 5AH as you will have 4 extra battery connection
leads to deal with and connect correctly. I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the "CAL" light goes
out when AC power is not applied. So shipping to Cal Lab can be a problem
if it is a distance away, or you have to use a shipper like UPS or FEDEX
and
you ship the night before and then use their "Morning delivery" and the Cal
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the "ext
power" plug to last longer. I have seen it done. The issue is to get the
Cal Lab to charge the extra battery before they ship the 732A back to you.
When you remove and work on the battery pack always have the AC power
plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply (battery)
voltage drops below about 21 volts the "CAL" light will go out. Below that
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the Reference
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification. When
the semiconductor junctions are unbiased and cool off when power is lost,
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is lost
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this. If
you don't have "low thermal" connection leads you can experience uV changes
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads. I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A. Using the MATH function and all of the data you
can collect. That was after waiting for several minutes after plugging in
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree on
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will only
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the capacitors. I
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once these
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs new
batteries. Any suggestions for sources (I haven't opened up the unit
I want to make sure it works before doing that). Also received the
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
and follow the instructions there.
Bill,
I have never been able to use banana plugs of any kind, with my 3458A,
due to their excessive thermal mass. I use 24 gauge wires to make a
twisted pair by hand with no connectors for precision voltage work. As
brought out in an earlier post by acbern, it doesn't matter about
plating or solder at least as long as your twisted pairs are made from
the same wire so the thermocouples balance out.
I am a big fan of M27500 shielded twisted pair wire for everyday use. It
is durable, heat resistant and capable of very low voltage use. The
insulator is Tefzel and the plating used to be silver but may be
something else now. It take a few minutes work for proper termination
but it can be done for free. I tend to use a lot of the 24 gauge M27500
STP wire since it is fast to come to temperature, flexible and durable
enough to last quite awhile.
eBay always has quite a few small lengths available and Sierra Nevada
Products has a bunch of it. Sierra Nevada Products says that it is
Teflon but I think it may be Tefzel. I can help with termination
instructions if wanted.
Charlie
On 8/24/2014 7:22 AM, Randy Evans wrote:
> Bill,
>
> I have convinced myself that the problem I an seeing is due to thermals.
> If I move the cables (with gold-plated banana plugs) using a small towel
> rather than letting my hand touch the plugs, it is much more stable. If I
> then hold the banana plug with my hand after the reading has stabilized,
> the reading drifts rapidly upward. I am trying to check the stability of
> the reading but I haven't figured out the MATH function yet. I assume this
> is a programmed function using GPIB only?
>
> The stability I am seeing by manually recording readings (using NLPC of 100
> and 1000) is much greater than what you are measuring on your system. Not
> sure how to ascertain if it's the 3458A or the 732. The value of the
> readings are very different between the two - the 3458 reads about 50 uV
> high on the 10 V output and about 12 uV low on the 1V output. Rather large
> differences (this is after an ACAL). I need to find some better cables to
> make sure the errors are not due to thermals again.
>
> Randy
>
>
> On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
>
>> Randy:
>>
>> I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
>> perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
>> others
>> that are in this size and package. Order from one of the usual electronics
>> distributors like Allied, Mouser, Digikey. This is a very common battery
>> as
>> it is used in a lot of "EXIT" signs so they are lighted when the power goes
>> out. I don't see how the 12V 7AH will fit as they are too large. I guess
>> you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
>> but the terminals are in the wrong place so you will have to "nibble" out
>> the aluminum plate that holds them in the 732A battery pack. You have to
>> be
>> careful if you use the 12v 5AH as you will have 4 extra battery connection
>> leads to deal with and connect correctly. I would stick with the 6V 4AH.
>> New batteries will last around 12 to 14 hours before the "CAL" light goes
>> out when AC power is not applied. So shipping to Cal Lab can be a problem
>> if it is a distance away, or you have to use a shipper like UPS or FEDEX
>> and
>> you ship the night before and then use their "Morning delivery" and the Cal
>> Lab is expecting your 732A. Same on the way back to you. Of course you
>> could always strap another battery on the 732A and hook it up to the "ext
>> power" plug to last longer. I have seen it done. The issue is to get the
>> Cal Lab to charge the extra battery before they ship the 732A back to you.
>>
>> When you remove and work on the battery pack always have the AC power
>> plugged in. The "CAL" led will stay on because the 18.6 v regulated supply
>> is working.
>>
>> The "CAL" light is to indicate that power has not been lost to the
>> Reference Amp or other associated circuits. When the raw supply (battery)
>> voltage drops below about 21 volts the "CAL" light will go out. Below that
>> voltage the heater circuits will not work correctly and the 18.6 volt
>> regulated supply will not regulate. The requirement is that the Reference
>> Amp be kept "alive" at all times to maintain the output voltage that was
>> measured at the time of the most recent Calibration or Certification. When
>> the semiconductor junctions are unbiased and cool off when power is lost,
>> and then power is restored the result will be a different 10 volts than
>> before the power failure. My experience is that after all of the years
>> that
>> these units have been powered up, this won't happen and when power is lost
>> and then restored, even months later, the 732A will come back to almost
>> exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
>> after 24 hours of "warm up".
>>
>> What type of hookup leads are you using when measuring the 1 volt
>> output? What is the PLC set to? I always use 100 PLC to measure this. If
>> you don't have "low thermal" connection leads you can experience uV changes
>> for a minute or more after plugging in the leads due to the "thermals"
>> generated because of the difference in temperature between the banana jacks
>> on the 732A and the banana plugs of the leads. I have found that even just
>> plugging in the lead will generate a thermal difference because of
>> difference of temps and some heating due to the physical act of just
>> inserting the plug because of friction between the jack and plug (my theory
>> at any rate). You have to allow at least a minute or more before being
>> able
>> to make a measurement after plugging in the leads. I just measured the
>> variation of the 1 volt output of my 732A and using my 3458A and I got a
>> total difference of 0.159 uV over 40 measurements using 100 PLC on the 1
>> volt range of the 3458A. Using the MATH function and all of the data you
>> can collect. That was after waiting for several minutes after plugging in
>> the leads.
>>
>> I hope all of this helps.
>>
>> Bill
>>
>>
>>
>> ----- Original Message -----
>> From: "Randy Evans" <randyevans2688@gmail.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Sent: Friday, August 22, 2014 9:03 PM
>> Subject: Re: [volt-nuts] 732A and Prologix received
>>
>>
>>> Todd,
>>>
>>> Thanks for the info. I have several Panasonic 12V 7 AH batteries that I
>>> keep topped off and they have very low current draw (~2 to 3 mA at 13.5
>>> VDC) when charged and at their float voltage, so I am pretty sure they
>> are
>>> in good condition. I will look at getting those in the units after I
>>> ascertain the condition of the 732.
>>>
>>> So now I have a what appears to be a functioning 3458A and a 732A but
>> they
>>> slightly disagree. I am like the man with two watches that disagree on
>> the
>>> time - which is correct? For the moment, i am only concerned with
>>> stability. The need for absolute accuracy will come later.
>>>
>>> Randy
>>>
>>>
>>> On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
>> wrote:
>>>> Randy,
>>>>
>>>> You have two possible choices. It can be configured with 4 x 6v 4Ah
>>>> batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
>>>> modified the battery pack already. A couple of mine needed a nibbler
>> tool
>>>> to remove enough of the aluminum cover that fits over the tops of the
>>>> batteries. The original cover will short out to the battery tabs
>> regardless
>>>> of the battery configuration if this is not done.
>>>>
>>>> You can find larger capacity batteries that will give you slightly more
>>>> battery life. I lost a couple sets of mail-order batteries after a few
>>>> extended outages. I would recommend going with locally bought batteries
>>>> instead of the cheaper mail order. My local Batteries Plus will
>> typically
>>>> have some warranty if I remember correctly. Moving forward I will only
>> use
>>>> 2 12v batteries and pre-charge them on a battery charger to equalize
>> them
>>>> before putting them in the 732A. I think the cheap batteries did not
>>>> discharge equally, and would not recover when power was applied.
>>>>
>>>> Inspect the back plane for damaged traces and look at the capacitors. I
>> had
>>>> a few that looked questionable. So far, I have replaced all the big
>> caps
>> on
>>>> the pre-regulator and regulator boards. My feeling is that once these
>> go
>>>> online, they should run as long as possible between repairs.
>>>>
>>>> The battery charger circuit may need adjusting. I tweaked mine and it
>>>> seemed to work fine.
>>>>
>>>> Todd
>>>>
>>>>
>>>> On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
>> randyevans2688@gmail.com>
>>>> wrote:
>>>>
>>>>> I received my Fluke 732A today. Just powered it up but it needs new
>>>>> batteries. Any suggestions for sources (I haven't opened up the unit
>>>> yet -
>>>>> I want to make sure it works before doing that). Also received the
>>>>> ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran data
>>>> dumper
>>>>> program to get the CAL data from my 3458A. Should be a busy weekend.
>>>>>
>>>>> Randy
>>>>> _______________________________________________
>>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>>> To unsubscribe, go to
>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>>> and follow the instructions there.
>>>>>
>>>> _______________________________________________
>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>> and follow the instructions there.
>>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
>
D
Don@True-Cal
Sun, Aug 24, 2014 8:46 PM
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
As you observed, merely the friction of plugging in the spring banana plug
and heat transfer from your fingers will require ~3-5 minutes to stabilize.
Low mass terminals help with the time to initially stabilize but is not best
if dealing with local air current drafts that upset the instantaneous
thermal gradients between the higher mass device terminal and the lower mass
lead terminal. This issue is a trade-off and is specifically considered in
the Fluke cables but being careful with localized drafts around the
terminals during critical measurements can virtually eliminate this error
source.
I have attached an ~30-hour measurement run that I consistently do that
should give you some idea of expected measurement drift over time. Virtually
all of the measurement drift is due to the 3458A internal temperature
differences around the 7V reference caused by ambient temperature change.
The drift associated with the 732A is probably about 2-magnitudes less at
this ambient temp drift. In my situation, the inferred ambient temperature
is cycling with the home air-conditioning. Note the initial calibration
temperatures as well as the ACal temperatures and where the 3458A is
measuring exactly 10V. My primary 732A has been powered without loss for 4
years and >5 years before that. The 3458A is Agilent (Loveland) Cal'ed
yearly and is powered 24-7-365 except for the occasional mains power loss.
The graphical measurements is using a homegrown Agilent VEE program.
Don Johnson
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Randy Evans
Sent: Sunday, August 24, 2014 9:22 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If I
then hold the banana plug with my hand after the reading has stabilized, the
reading drifts rapidly upward. I am trying to check the stability of the
reading but I haven't figured out the MATH function yet. I assume this is a
programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of 100
and 1000) is much greater than what you are measuring on your system. Not
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather large
differences (this is after an ACAL). I need to find some better cables to
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others that are in this size and package. Order from one of the usual
electronics distributors like Allied, Mouser, Digikey. This is a very
common battery as it is used in a lot of "EXIT" signs so they are
lighted when the power goes out. I don't see how the 12V 7AH will fit
as they are too large. I guess you could use a 12V 5AH (PS1250) as it
is the same size as 2 x 6 volt 4 AH but the terminals are in the wrong
place so you will have to "nibble" out the aluminum plate that holds
them in the 732A battery pack. You have to be careful if you use the
12v 5AH as you will have 4 extra battery connection leads to deal with
and connect correctly. I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the "CAL" light
goes out when AC power is not applied. So shipping to Cal Lab can be
a problem if it is a distance away, or you have to use a shipper like
UPS or FEDEX and you ship the night before and then use their "Morning
delivery" and the Cal Lab is expecting your 732A. Same on the way
back to you. Of course you could always strap another battery on the
732A and hook it up to the "ext power" plug to last longer. I have
seen it done. The issue is to get the Cal Lab to charge the extra
battery before they ship the 732A back to you.
When you remove and work on the battery pack always have the AC
power plugged in. The "CAL" led will stay on because the 18.6 v
regulated supply is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
(battery) voltage drops below about 21 volts the "CAL" light will go
out. Below that voltage the heater circuits will not work correctly
and the 18.6 volt regulated supply will not regulate. The requirement
is that the Reference Amp be kept "alive" at all times to maintain the
output voltage that was measured at the time of the most recent
Calibration or Certification. When the semiconductor junctions are
unbiased and cool off when power is lost, and then power is restored
the result will be a different 10 volts than before the power failure.
My experience is that after all of the years that these units have
been powered up, this won't happen and when power is lost and then
restored, even months later, the 732A will come back to almost exactly
the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure
this. If you don't have "low thermal" connection leads you can
experience uV changes for a minute or more after plugging in the leads due
generated because of the difference in temperature between the banana
jacks on the 732A and the banana plugs of the leads. I have found
that even just plugging in the lead will generate a thermal difference
because of difference of temps and some heating due to the physical
act of just inserting the plug because of friction between the jack
and plug (my theory at any rate). You have to allow at least a minute
or more before being able to make a measurement after plugging in the
leads. I just measured the variation of the 1 volt output of my 732A
and using my 3458A and I got a total difference of 0.159 uV over 40
measurements using 100 PLC on the 1 volt range of the 3458A. Using
the MATH function and all of the data you can collect. That was after
waiting for several minutes after plugging in the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
that I keep topped off and they have very low current draw (~2 to 3
mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure
they
in good condition. I will look at getting those in the units after
I ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
but
slightly disagree. I am like the man with two watches that disagree
on
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v
4Ah batteries or 2 x 12v 7Ah batteries. Hopefully the previous
owner has modified the battery pack already. A couple of mine
needed a nibbler
to remove enough of the aluminum cover that fits over the tops of
the batteries. The original cover will short out to the battery
tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
more battery life. I lost a couple sets of mail-order batteries
after a few extended outages. I would recommend going with locally
bought batteries instead of the cheaper mail order. My local
Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
only
2 12v batteries and pre-charge them on a battery charger to
equalize
before putting them in the 732A. I think the cheap batteries did
not discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
capacitors. I
a few that looked questionable. So far, I have replaced all the
big
the pre-regulator and regulator boards. My feeling is that once
these
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
it seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
new batteries. Any suggestions for sources (I haven't opened up
the unit
I want to make sure it works before doing that). Also received
the ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
ran data
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
As you observed, merely the friction of plugging in the spring banana plug
and heat transfer from your fingers will require ~3-5 minutes to stabilize.
Low mass terminals help with the time to initially stabilize but is not best
if dealing with local air current drafts that upset the instantaneous
thermal gradients between the higher mass device terminal and the lower mass
lead terminal. This issue is a trade-off and is specifically considered in
the Fluke cables but being careful with localized drafts around the
terminals during critical measurements can virtually eliminate this error
source.
I have attached an ~30-hour measurement run that I consistently do that
should give you some idea of expected measurement drift over time. Virtually
all of the measurement drift is due to the 3458A internal temperature
differences around the 7V reference caused by ambient temperature change.
The drift associated with the 732A is probably about 2-magnitudes less at
this ambient temp drift. In my situation, the inferred ambient temperature
is cycling with the home air-conditioning. Note the initial calibration
temperatures as well as the ACal temperatures and where the 3458A is
measuring exactly 10V. My primary 732A has been powered without loss for 4
years and >5 years before that. The 3458A is Agilent (Loveland) Cal'ed
yearly and is powered 24-7-365 except for the occasional mains power loss.
The graphical measurements is using a homegrown Agilent VEE program.
Don Johnson
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Randy Evans
Sent: Sunday, August 24, 2014 9:22 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If I
then hold the banana plug with my hand after the reading has stabilized, the
reading drifts rapidly upward. I am trying to check the stability of the
reading but I haven't figured out the MATH function yet. I assume this is a
programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of 100
and 1000) is much greater than what you are measuring on your system. Not
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather large
differences (this is after an ACAL). I need to find some better cables to
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
> Randy:
>
> I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> others that are in this size and package. Order from one of the usual
> electronics distributors like Allied, Mouser, Digikey. This is a very
> common battery as it is used in a lot of "EXIT" signs so they are
> lighted when the power goes out. I don't see how the 12V 7AH will fit
> as they are too large. I guess you could use a 12V 5AH (PS1250) as it
> is the same size as 2 x 6 volt 4 AH but the terminals are in the wrong
> place so you will have to "nibble" out the aluminum plate that holds
> them in the 732A battery pack. You have to be careful if you use the
> 12v 5AH as you will have 4 extra battery connection leads to deal with
> and connect correctly. I would stick with the 6V 4AH.
> New batteries will last around 12 to 14 hours before the "CAL" light
> goes out when AC power is not applied. So shipping to Cal Lab can be
> a problem if it is a distance away, or you have to use a shipper like
> UPS or FEDEX and you ship the night before and then use their "Morning
> delivery" and the Cal Lab is expecting your 732A. Same on the way
> back to you. Of course you could always strap another battery on the
> 732A and hook it up to the "ext power" plug to last longer. I have
> seen it done. The issue is to get the Cal Lab to charge the extra
> battery before they ship the 732A back to you.
>
> When you remove and work on the battery pack always have the AC
> power plugged in. The "CAL" led will stay on because the 18.6 v
> regulated supply is working.
>
> The "CAL" light is to indicate that power has not been lost to the
> Reference Amp or other associated circuits. When the raw supply
> (battery) voltage drops below about 21 volts the "CAL" light will go
> out. Below that voltage the heater circuits will not work correctly
> and the 18.6 volt regulated supply will not regulate. The requirement
> is that the Reference Amp be kept "alive" at all times to maintain the
> output voltage that was measured at the time of the most recent
> Calibration or Certification. When the semiconductor junctions are
> unbiased and cool off when power is lost, and then power is restored
> the result will be a different 10 volts than before the power failure.
> My experience is that after all of the years that these units have
> been powered up, this won't happen and when power is lost and then
> restored, even months later, the 732A will come back to almost exactly
> the same 10 volts as when they lost power, usually with in 0.2 PPM
> after 24 hours of "warm up".
>
> What type of hookup leads are you using when measuring the 1 volt
> output? What is the PLC set to? I always use 100 PLC to measure
> this. If you don't have "low thermal" connection leads you can
> experience uV changes for a minute or more after plugging in the leads due
to the "thermals"
> generated because of the difference in temperature between the banana
> jacks on the 732A and the banana plugs of the leads. I have found
> that even just plugging in the lead will generate a thermal difference
> because of difference of temps and some heating due to the physical
> act of just inserting the plug because of friction between the jack
> and plug (my theory at any rate). You have to allow at least a minute
> or more before being able to make a measurement after plugging in the
> leads. I just measured the variation of the 1 volt output of my 732A
> and using my 3458A and I got a total difference of 0.159 uV over 40
> measurements using 100 PLC on the 1 volt range of the 3458A. Using
> the MATH function and all of the data you can collect. That was after
> waiting for several minutes after plugging in the leads.
>
> I hope all of this helps.
>
> Bill
>
>
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Friday, August 22, 2014 9:03 PM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Todd,
> >
> > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> > that I keep topped off and they have very low current draw (~2 to 3
> > mA at 13.5
> > VDC) when charged and at their float voltage, so I am pretty sure
> > they
> are
> > in good condition. I will look at getting those in the units after
> > I ascertain the condition of the 732.
> >
> > So now I have a what appears to be a functioning 3458A and a 732A
> > but
> they
> > slightly disagree. I am like the man with two watches that disagree
> > on
> the
> > time - which is correct? For the moment, i am only concerned with
> > stability. The need for absolute accuracy will come later.
> >
> > Randy
> >
> >
> > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> wrote:
> >
> > > Randy,
> > >
> > > You have two possible choices. It can be configured with 4 x 6v
> > > 4Ah batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> > > owner has modified the battery pack already. A couple of mine
> > > needed a nibbler
> tool
> > > to remove enough of the aluminum cover that fits over the tops of
> > > the batteries. The original cover will short out to the battery
> > > tabs
> regardless
> > > of the battery configuration if this is not done.
> > >
> > > You can find larger capacity batteries that will give you slightly
> > > more battery life. I lost a couple sets of mail-order batteries
> > > after a few extended outages. I would recommend going with locally
> > > bought batteries instead of the cheaper mail order. My local
> > > Batteries Plus will
> typically
> > > have some warranty if I remember correctly. Moving forward I will
> > > only
> use
> > > 2 12v batteries and pre-charge them on a battery charger to
> > > equalize
> them
> > > before putting them in the 732A. I think the cheap batteries did
> > > not discharge equally, and would not recover when power was applied.
> > >
> > > Inspect the back plane for damaged traces and look at the
> > > capacitors. I
> had
> > > a few that looked questionable. So far, I have replaced all the
> > > big
> caps
> on
> > > the pre-regulator and regulator boards. My feeling is that once
> > > these
> go
> > > online, they should run as long as possible between repairs.
> > >
> > > The battery charger circuit may need adjusting. I tweaked mine and
> > > it seemed to work fine.
> > >
> > > Todd
> > >
> > >
> > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> randyevans2688@gmail.com>
> > > wrote:
> > >
> > > > I received my Fluke 732A today. Just powered it up but it needs
> > > > new batteries. Any suggestions for sources (I haven't opened up
> > > > the unit
> > > yet -
> > > > I want to make sure it works before doing that). Also received
> > > > the ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
> > > > ran data
> > > dumper
> > > > program to get the CAL data from my 3458A. Should be a busy
weekend.
> > > >
> > > > Randy
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go
> > > > to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
BG
Bill Gold
Sun, Aug 24, 2014 9:44 PM
Todd:
I check the batteries every month per the 732A manual. Usually I keep a
few used cells from previous packs around that seem to be ok after checking
them for capacity just like checking the battery pack. I can usually spot
cells going bad with that method. If I don't have a spare set then I just
substitute a used cell until I receive a new set.
I agree with you that the cheap ones are crap. I have used Power Sonic
a lot with some a few EaglePicher and one set of Enersys. The EaglePicher
are crap so I will never use them. EaglePicher have had a couple out of 8
that have failed very early. I have never tried Panasonic. So I stick with
the Power Sonic's. I am not sure why the 12 volt ones would be better than
the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3
lead acid cells. If one of the 6 cells in the 12 volt battery go bad then
the whole battery has to be changed. So you better buy good quality 12 volt
batteries.
I just can't seem to source batteries locally easily at a reasonable
price. So I order them from Allied usually.
Bill
----- Original Message -----
From: "Todd Micallef" tmicallef@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 8:13 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes,
I was mistaken. The 12V 5ah batteries are the ones that I am using in one
of my 732A. There is more play inside the tray with the 12v batteries by
several mm as compared to the 6v which only has a 2-3mm. The 12v
is not difficult, but it is easier if the battery tray has already been
machined for the 12v batteries. The battery tabs don't line up well with
the existing holes and need to be widened. Two additional holes must also
be added.
Sourcing the batteries locally is more of a convenience than waiting for
the delivery. The el-cheapo 6v batteries are a waste of money and I have 8
batteries so far to prove it. They died the first time I had a 6hr outage
and would not hold a charge after that.
Todd
On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or FEDEX
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure they
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Todd:
I check the batteries every month per the 732A manual. Usually I keep a
few used cells from previous packs around that seem to be ok after checking
them for capacity just like checking the battery pack. I can usually spot
cells going bad with that method. If I don't have a spare set then I just
substitute a used cell until I receive a new set.
I agree with you that the cheap ones are crap. I have used Power Sonic
a lot with some a few EaglePicher and one set of Enersys. The EaglePicher
are crap so I will never use them. EaglePicher have had a couple out of 8
that have failed very early. I have never tried Panasonic. So I stick with
the Power Sonic's. I am not sure why the 12 volt ones would be better than
the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3
lead acid cells. If one of the 6 cells in the 12 volt battery go bad then
the whole battery has to be changed. So you better buy good quality 12 volt
batteries.
I just can't seem to source batteries locally easily at a reasonable
price. So I order them from Allied usually.
Bill
----- Original Message -----
From: "Todd Micallef" <tmicallef@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Sunday, August 24, 2014 8:13 AM
Subject: Re: [volt-nuts] 732A and Prologix received
> Bill,
>
> I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes,
> I was mistaken. The 12V 5ah batteries are the ones that I am using in one
> of my 732A. There is more play inside the tray with the 12v batteries by
> several mm as compared to the 6v which only has a 2-3mm. The 12v
conversion
> is not difficult, but it is easier if the battery tray has already been
> machined for the 12v batteries. The battery tabs don't line up well with
> the existing holes and need to be widened. Two additional holes must also
> be added.
>
> Sourcing the batteries locally is more of a convenience than waiting for
> the delivery. The el-cheapo 6v batteries are a waste of money and I have 8
> batteries so far to prove it. They died the first time I had a 6hr outage
> and would not hold a charge after that.
>
> Todd
>
>
>
>
> On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold <wpgold3637@att.net> wrote:
>
> > Randy:
> >
> > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> > others
> > that are in this size and package. Order from one of the usual
electronics
> > distributors like Allied, Mouser, Digikey. This is a very common
battery
> > as
> > it is used in a lot of "EXIT" signs so they are lighted when the power
goes
> > out. I don't see how the 12V 7AH will fit as they are too large. I
guess
> > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
AH
> > but the terminals are in the wrong place so you will have to "nibble"
out
> > the aluminum plate that holds them in the 732A battery pack. You have
to
> > be
> > careful if you use the 12v 5AH as you will have 4 extra battery
connection
> > leads to deal with and connect correctly. I would stick with the 6V
4AH.
> > New batteries will last around 12 to 14 hours before the "CAL" light
goes
> > out when AC power is not applied. So shipping to Cal Lab can be a
problem
> > if it is a distance away, or you have to use a shipper like UPS or FEDEX
> > and
> > you ship the night before and then use their "Morning delivery" and the
Cal
> > Lab is expecting your 732A. Same on the way back to you. Of course you
> > could always strap another battery on the 732A and hook it up to the
"ext
> > power" plug to last longer. I have seen it done. The issue is to get
the
> > Cal Lab to charge the extra battery before they ship the 732A back to
you.
> >
> > When you remove and work on the battery pack always have the AC
power
> > plugged in. The "CAL" led will stay on because the 18.6 v regulated
supply
> > is working.
> >
> > The "CAL" light is to indicate that power has not been lost to the
> > Reference Amp or other associated circuits. When the raw supply
(battery)
> > voltage drops below about 21 volts the "CAL" light will go out. Below
that
> > voltage the heater circuits will not work correctly and the 18.6 volt
> > regulated supply will not regulate. The requirement is that the
Reference
> > Amp be kept "alive" at all times to maintain the output voltage that was
> > measured at the time of the most recent Calibration or Certification.
When
> > the semiconductor junctions are unbiased and cool off when power is
lost,
> > and then power is restored the result will be a different 10 volts than
> > before the power failure. My experience is that after all of the years
> > that
> > these units have been powered up, this won't happen and when power is
lost
> > and then restored, even months later, the 732A will come back to almost
> > exactly the same 10 volts as when they lost power, usually with in 0.2
PPM
> > after 24 hours of "warm up".
> >
> > What type of hookup leads are you using when measuring the 1 volt
> > output? What is the PLC set to? I always use 100 PLC to measure this.
If
> > you don't have "low thermal" connection leads you can experience uV
changes
> > for a minute or more after plugging in the leads due to the "thermals"
> > generated because of the difference in temperature between the banana
jacks
> > on the 732A and the banana plugs of the leads. I have found that even
just
> > plugging in the lead will generate a thermal difference because of
> > difference of temps and some heating due to the physical act of just
> > inserting the plug because of friction between the jack and plug (my
theory
> > at any rate). You have to allow at least a minute or more before being
> > able
> > to make a measurement after plugging in the leads. I just measured the
> > variation of the 1 volt output of my 732A and using my 3458A and I got a
> > total difference of 0.159 uV over 40 measurements using 100 PLC on the
1
> > volt range of the 3458A. Using the MATH function and all of the data
you
> > can collect. That was after waiting for several minutes after plugging
in
> > the leads.
> >
> > I hope all of this helps.
> >
> > Bill
> >
> >
> >
> > ----- Original Message -----
> > From: "Randy Evans" <randyevans2688@gmail.com>
> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Sent: Friday, August 22, 2014 9:03 PM
> > Subject: Re: [volt-nuts] 732A and Prologix received
> >
> >
> > > Todd,
> > >
> > > Thanks for the info. I have several Panasonic 12V 7 AH batteries that
I
> > > keep topped off and they have very low current draw (~2 to 3 mA at
13.5
> > > VDC) when charged and at their float voltage, so I am pretty sure they
> > are
> > > in good condition. I will look at getting those in the units after I
> > > ascertain the condition of the 732.
> > >
> > > So now I have a what appears to be a functioning 3458A and a 732A but
> > they
> > > slightly disagree. I am like the man with two watches that disagree
on
> > the
> > > time - which is correct? For the moment, i am only concerned with
> > > stability. The need for absolute accuracy will come later.
> > >
> > > Randy
> > >
> > >
> > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> > wrote:
> > >
> > > > Randy,
> > > >
> > > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> > > > modified the battery pack already. A couple of mine needed a nibbler
> > tool
> > > > to remove enough of the aluminum cover that fits over the tops of
the
> > > > batteries. The original cover will short out to the battery tabs
> > regardless
> > > > of the battery configuration if this is not done.
> > > >
> > > > You can find larger capacity batteries that will give you slightly
more
> > > > battery life. I lost a couple sets of mail-order batteries after a
few
> > > > extended outages. I would recommend going with locally bought
batteries
> > > > instead of the cheaper mail order. My local Batteries Plus will
> > typically
> > > > have some warranty if I remember correctly. Moving forward I will
only
> > use
> > > > 2 12v batteries and pre-charge them on a battery charger to equalize
> > them
> > > > before putting them in the 732A. I think the cheap batteries did not
> > > > discharge equally, and would not recover when power was applied.
> > > >
> > > > Inspect the back plane for damaged traces and look at the
capacitors. I
> > had
> > > > a few that looked questionable. So far, I have replaced all the big
> > caps
> > on
> > > > the pre-regulator and regulator boards. My feeling is that once
these
> > go
> > > > online, they should run as long as possible between repairs.
> > > >
> > > > The battery charger circuit may need adjusting. I tweaked mine and
it
> > > > seemed to work fine.
> > > >
> > > > Todd
> > > >
> > > >
> > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > randyevans2688@gmail.com>
> > > > wrote:
> > > >
> > > > > I received my Fluke 732A today. Just powered it up but it needs
new
> > > > > batteries. Any suggestions for sources (I haven't opened up the
unit
> > > > yet -
> > > > > I want to make sure it works before doing that). Also received
the
> > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
data
> > > > dumper
> > > > > program to get the CAL data from my 3458A. Should be a busy
weekend.
> > > > >
> > > > > Randy
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
BG
Bill Gold
Sun, Aug 24, 2014 11:46 PM
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
the instability of the readings it is hard to determine which is causing the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and then
trigger the sequence of 40 readings when I push the "ENTER" button. You can
do all of this manually from the keypads but since I use this sequence a lot
I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts. This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above. I would
use this to determine where your problem might exist. Just having the meter
input shorted will point you in the right direction. Meter, cables or 732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If I
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system. Not
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables to
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or FEDEX
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure they
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
the instability of the readings it is hard to determine which is causing the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and then
trigger the sequence of 40 readings when I push the "ENTER" button. You can
do all of this manually from the keypads but since I use this sequence a lot
I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts. This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above. I would
use this to determine where your problem might exist. Just having the meter
input shorted will point you in the right direction. Meter, cables or 732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
> Bill,
>
> I have convinced myself that the problem I an seeing is due to thermals.
> If I move the cables (with gold-plated banana plugs) using a small towel
> rather than letting my hand touch the plugs, it is much more stable. If I
> then hold the banana plug with my hand after the reading has stabilized,
> the reading drifts rapidly upward. I am trying to check the stability of
> the reading but I haven't figured out the MATH function yet. I assume
this
> is a programmed function using GPIB only?
>
> The stability I am seeing by manually recording readings (using NLPC of
100
> and 1000) is much greater than what you are measuring on your system. Not
> sure how to ascertain if it's the 3458A or the 732. The value of the
> readings are very different between the two - the 3458 reads about 50 uV
> high on the 10 V output and about 12 uV low on the 1V output. Rather
large
> differences (this is after an ACAL). I need to find some better cables to
> make sure the errors are not due to thermals again.
>
> Randy
>
>
> On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
>
> > Randy:
> >
> > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> > others
> > that are in this size and package. Order from one of the usual
electronics
> > distributors like Allied, Mouser, Digikey. This is a very common
battery
> > as
> > it is used in a lot of "EXIT" signs so they are lighted when the power
goes
> > out. I don't see how the 12V 7AH will fit as they are too large. I
guess
> > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
AH
> > but the terminals are in the wrong place so you will have to "nibble"
out
> > the aluminum plate that holds them in the 732A battery pack. You have
to
> > be
> > careful if you use the 12v 5AH as you will have 4 extra battery
connection
> > leads to deal with and connect correctly. I would stick with the 6V
4AH.
> > New batteries will last around 12 to 14 hours before the "CAL" light
goes
> > out when AC power is not applied. So shipping to Cal Lab can be a
problem
> > if it is a distance away, or you have to use a shipper like UPS or FEDEX
> > and
> > you ship the night before and then use their "Morning delivery" and the
Cal
> > Lab is expecting your 732A. Same on the way back to you. Of course you
> > could always strap another battery on the 732A and hook it up to the
"ext
> > power" plug to last longer. I have seen it done. The issue is to get
the
> > Cal Lab to charge the extra battery before they ship the 732A back to
you.
> >
> > When you remove and work on the battery pack always have the AC
power
> > plugged in. The "CAL" led will stay on because the 18.6 v regulated
supply
> > is working.
> >
> > The "CAL" light is to indicate that power has not been lost to the
> > Reference Amp or other associated circuits. When the raw supply
(battery)
> > voltage drops below about 21 volts the "CAL" light will go out. Below
that
> > voltage the heater circuits will not work correctly and the 18.6 volt
> > regulated supply will not regulate. The requirement is that the
Reference
> > Amp be kept "alive" at all times to maintain the output voltage that was
> > measured at the time of the most recent Calibration or Certification.
When
> > the semiconductor junctions are unbiased and cool off when power is
lost,
> > and then power is restored the result will be a different 10 volts than
> > before the power failure. My experience is that after all of the years
> > that
> > these units have been powered up, this won't happen and when power is
lost
> > and then restored, even months later, the 732A will come back to almost
> > exactly the same 10 volts as when they lost power, usually with in 0.2
PPM
> > after 24 hours of "warm up".
> >
> > What type of hookup leads are you using when measuring the 1 volt
> > output? What is the PLC set to? I always use 100 PLC to measure this.
If
> > you don't have "low thermal" connection leads you can experience uV
changes
> > for a minute or more after plugging in the leads due to the "thermals"
> > generated because of the difference in temperature between the banana
jacks
> > on the 732A and the banana plugs of the leads. I have found that even
just
> > plugging in the lead will generate a thermal difference because of
> > difference of temps and some heating due to the physical act of just
> > inserting the plug because of friction between the jack and plug (my
theory
> > at any rate). You have to allow at least a minute or more before being
> > able
> > to make a measurement after plugging in the leads. I just measured the
> > variation of the 1 volt output of my 732A and using my 3458A and I got a
> > total difference of 0.159 uV over 40 measurements using 100 PLC on the
1
> > volt range of the 3458A. Using the MATH function and all of the data
you
> > can collect. That was after waiting for several minutes after plugging
in
> > the leads.
> >
> > I hope all of this helps.
> >
> > Bill
> >
> >
> >
> > ----- Original Message -----
> > From: "Randy Evans" <randyevans2688@gmail.com>
> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Sent: Friday, August 22, 2014 9:03 PM
> > Subject: Re: [volt-nuts] 732A and Prologix received
> >
> >
> > > Todd,
> > >
> > > Thanks for the info. I have several Panasonic 12V 7 AH batteries that
I
> > > keep topped off and they have very low current draw (~2 to 3 mA at
13.5
> > > VDC) when charged and at their float voltage, so I am pretty sure they
> > are
> > > in good condition. I will look at getting those in the units after I
> > > ascertain the condition of the 732.
> > >
> > > So now I have a what appears to be a functioning 3458A and a 732A but
> > they
> > > slightly disagree. I am like the man with two watches that disagree
on
> > the
> > > time - which is correct? For the moment, i am only concerned with
> > > stability. The need for absolute accuracy will come later.
> > >
> > > Randy
> > >
> > >
> > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> > wrote:
> > >
> > > > Randy,
> > > >
> > > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
> > > > modified the battery pack already. A couple of mine needed a nibbler
> > tool
> > > > to remove enough of the aluminum cover that fits over the tops of
the
> > > > batteries. The original cover will short out to the battery tabs
> > regardless
> > > > of the battery configuration if this is not done.
> > > >
> > > > You can find larger capacity batteries that will give you slightly
more
> > > > battery life. I lost a couple sets of mail-order batteries after a
few
> > > > extended outages. I would recommend going with locally bought
batteries
> > > > instead of the cheaper mail order. My local Batteries Plus will
> > typically
> > > > have some warranty if I remember correctly. Moving forward I will
only
> > use
> > > > 2 12v batteries and pre-charge them on a battery charger to equalize
> > them
> > > > before putting them in the 732A. I think the cheap batteries did not
> > > > discharge equally, and would not recover when power was applied.
> > > >
> > > > Inspect the back plane for damaged traces and look at the
capacitors. I
> > had
> > > > a few that looked questionable. So far, I have replaced all the big
> > caps
> > on
> > > > the pre-regulator and regulator boards. My feeling is that once
these
> > go
> > > > online, they should run as long as possible between repairs.
> > > >
> > > > The battery charger circuit may need adjusting. I tweaked mine and
it
> > > > seemed to work fine.
> > > >
> > > > Todd
> > > >
> > > >
> > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > randyevans2688@gmail.com>
> > > > wrote:
> > > >
> > > > > I received my Fluke 732A today. Just powered it up but it needs
new
> > > > > batteries. Any suggestions for sources (I haven't opened up the
unit
> > > > yet -
> > > > > I want to make sure it works before doing that). Also received
the
> > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
data
> > > > dumper
> > > > > program to get the CAL data from my 3458A. Should be a busy
weekend.
> > > > >
> > > > > Randy
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
TM
Todd Micallef
Sun, Aug 24, 2014 11:52 PM
Bill,
Thanks for the tip. Do you have an automated measurement system for the
10v inter-comparisons or do you manually log them with a spreadsheet?
I don't think the 12v are better, but it is interesting that some people
have modified them to dual batteries. I do not think there is any added
benefit unless someone happens to use a lot of those particular 12v
batteries.
So far I have two with the 6v packs and one with the 12v. I have a
fourth one waiting to be repaired once I finish my current projects.
Todd
Sent from my iPad
Todd:
I check the batteries every month per the 732A manual. Usually I keep
few used cells from previous packs around that seem to be ok after
them for capacity just like checking the battery pack. I can usually spot
cells going bad with that method. If I don't have a spare set then I just
substitute a used cell until I receive a new set.
I agree with you that the cheap ones are crap. I have used Power Sonic
a lot with some a few EaglePicher and one set of Enersys. The EaglePicher
are crap so I will never use them. EaglePicher have had a couple out of 8
that have failed very early. I have never tried Panasonic. So I stick
the Power Sonic's. I am not sure why the 12 volt ones would be better
the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are
lead acid cells. If one of the 6 cells in the 12 volt battery go bad then
the whole battery has to be changed. So you better buy good quality 12
batteries.
I just can't seem to source batteries locally easily at a reasonable
price. So I order them from Allied usually.
Bill
----- Original Message -----
From: "Todd Micallef" <tmicallef@gmail.com javascript:;>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com
Sent: Sunday, August 24, 2014 8:13 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But
I was mistaken. The 12V 5ah batteries are the ones that I am using in one
of my 732A. There is more play inside the tray with the 12v batteries by
several mm as compared to the 6v which only has a 2-3mm. The 12v
is not difficult, but it is easier if the battery tray has already been
machined for the 12v batteries. The battery tabs don't line up well with
the existing holes and need to be widened. Two additional holes must also
be added.
Sourcing the batteries locally is more of a convenience than waiting for
the delivery. The el-cheapo 6v batteries are a waste of money and I have
batteries so far to prove it. They died the first time I had a 6hr outage
and would not hold a charge after that.
Todd
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or FEDEX
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course you
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries that
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure they
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to equalize
before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Bill,
Thanks for the tip. Do you have an automated measurement system for the
10v inter-comparisons or do you manually log them with a spreadsheet?
I don't think the 12v are better, but it is interesting that some people
have modified them to dual batteries. I do not think there is any added
benefit unless someone happens to use a lot of those particular 12v
batteries.
So far I have two with the 6v packs and one with the 12v. I have a
fourth one waiting to be repaired once I finish my current projects.
Todd
Sent from my iPad
> On Aug 24, 2014, at 17:44, "Bill Gold" <wpgold3637@att.net <javascript:;>>
wrote:
>
> Todd:
>
> I check the batteries every month per the 732A manual. Usually I keep
a
> few used cells from previous packs around that seem to be ok after
checking
> them for capacity just like checking the battery pack. I can usually spot
> cells going bad with that method. If I don't have a spare set then I just
> substitute a used cell until I receive a new set.
>
> I agree with you that the cheap ones are crap. I have used Power Sonic
> a lot with some a few EaglePicher and one set of Enersys. The EaglePicher
> are crap so I will never use them. EaglePicher have had a couple out of 8
> that have failed very early. I have never tried Panasonic. So I stick
with
> the Power Sonic's. I am not sure why the 12 volt ones would be better
than
> the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are
3
> lead acid cells. If one of the 6 cells in the 12 volt battery go bad then
> the whole battery has to be changed. So you better buy good quality 12
volt
> batteries.
>
> I just can't seem to source batteries locally easily at a reasonable
> price. So I order them from Allied usually.
>
> Bill
>
> ----- Original Message -----
> From: "Todd Micallef" <tmicallef@gmail.com <javascript:;>>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com
<javascript:;>>
> Sent: Sunday, August 24, 2014 8:13 AM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
>> Bill,
>>
>> I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But
yes,
>> I was mistaken. The 12V 5ah batteries are the ones that I am using in one
>> of my 732A. There is more play inside the tray with the 12v batteries by
>> several mm as compared to the 6v which only has a 2-3mm. The 12v
> conversion
>> is not difficult, but it is easier if the battery tray has already been
>> machined for the 12v batteries. The battery tabs don't line up well with
>> the existing holes and need to be widened. Two additional holes must also
>> be added.
>>
>> Sourcing the batteries locally is more of a convenience than waiting for
>> the delivery. The el-cheapo 6v batteries are a waste of money and I have
8
>> batteries so far to prove it. They died the first time I had a 6hr outage
>> and would not hold a charge after that.
>>
>> Todd
>>
>>
>>
>>
>>> On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold <wpgold3637@att.net
<javascript:;>> wrote:
>>>
>>> Randy:
>>>
>>> I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
>>> perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
>>> others
>>> that are in this size and package. Order from one of the usual
> electronics
>>> distributors like Allied, Mouser, Digikey. This is a very common
> battery
>>> as
>>> it is used in a lot of "EXIT" signs so they are lighted when the power
> goes
>>> out. I don't see how the 12V 7AH will fit as they are too large. I
> guess
>>> you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
> AH
>>> but the terminals are in the wrong place so you will have to "nibble"
> out
>>> the aluminum plate that holds them in the 732A battery pack. You have
> to
>>> be
>>> careful if you use the 12v 5AH as you will have 4 extra battery
> connection
>>> leads to deal with and connect correctly. I would stick with the 6V
> 4AH.
>>> New batteries will last around 12 to 14 hours before the "CAL" light
> goes
>>> out when AC power is not applied. So shipping to Cal Lab can be a
> problem
>>> if it is a distance away, or you have to use a shipper like UPS or FEDEX
>>> and
>>> you ship the night before and then use their "Morning delivery" and the
> Cal
>>> Lab is expecting your 732A. Same on the way back to you. Of course you
>>> could always strap another battery on the 732A and hook it up to the
> "ext
>>> power" plug to last longer. I have seen it done. The issue is to get
> the
>>> Cal Lab to charge the extra battery before they ship the 732A back to
> you.
>>>
>>> When you remove and work on the battery pack always have the AC
> power
>>> plugged in. The "CAL" led will stay on because the 18.6 v regulated
> supply
>>> is working.
>>>
>>> The "CAL" light is to indicate that power has not been lost to the
>>> Reference Amp or other associated circuits. When the raw supply
> (battery)
>>> voltage drops below about 21 volts the "CAL" light will go out. Below
> that
>>> voltage the heater circuits will not work correctly and the 18.6 volt
>>> regulated supply will not regulate. The requirement is that the
> Reference
>>> Amp be kept "alive" at all times to maintain the output voltage that was
>>> measured at the time of the most recent Calibration or Certification.
> When
>>> the semiconductor junctions are unbiased and cool off when power is
> lost,
>>> and then power is restored the result will be a different 10 volts than
>>> before the power failure. My experience is that after all of the years
>>> that
>>> these units have been powered up, this won't happen and when power is
> lost
>>> and then restored, even months later, the 732A will come back to almost
>>> exactly the same 10 volts as when they lost power, usually with in 0.2
> PPM
>>> after 24 hours of "warm up".
>>>
>>> What type of hookup leads are you using when measuring the 1 volt
>>> output? What is the PLC set to? I always use 100 PLC to measure this.
> If
>>> you don't have "low thermal" connection leads you can experience uV
> changes
>>> for a minute or more after plugging in the leads due to the "thermals"
>>> generated because of the difference in temperature between the banana
> jacks
>>> on the 732A and the banana plugs of the leads. I have found that even
> just
>>> plugging in the lead will generate a thermal difference because of
>>> difference of temps and some heating due to the physical act of just
>>> inserting the plug because of friction between the jack and plug (my
> theory
>>> at any rate). You have to allow at least a minute or more before being
>>> able
>>> to make a measurement after plugging in the leads. I just measured the
>>> variation of the 1 volt output of my 732A and using my 3458A and I got a
>>> total difference of 0.159 uV over 40 measurements using 100 PLC on the
> 1
>>> volt range of the 3458A. Using the MATH function and all of the data
> you
>>> can collect. That was after waiting for several minutes after plugging
> in
>>> the leads.
>>>
>>> I hope all of this helps.
>>>
>>> Bill
>>>
>>>
>>>
>>> ----- Original Message -----
>>> From: "Randy Evans" <randyevans2688@gmail.com <javascript:;>>
>>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com
<javascript:;>>
>>> Sent: Friday, August 22, 2014 9:03 PM
>>> Subject: Re: [volt-nuts] 732A and Prologix received
>>>
>>>
>>>> Todd,
>>>>
>>>> Thanks for the info. I have several Panasonic 12V 7 AH batteries that
> I
>>>> keep topped off and they have very low current draw (~2 to 3 mA at
> 13.5
>>>> VDC) when charged and at their float voltage, so I am pretty sure they
>>> are
>>>> in good condition. I will look at getting those in the units after I
>>>> ascertain the condition of the 732.
>>>>
>>>> So now I have a what appears to be a functioning 3458A and a 732A but
>>> they
>>>> slightly disagree. I am like the man with two watches that disagree
> on
>>> the
>>>> time - which is correct? For the moment, i am only concerned with
>>>> stability. The need for absolute accuracy will come later.
>>>>
>>>> Randy
>>>>
>>>>
>>>>> On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
<javascript:;>>
>>>> wrote:
>>>>
>>>>> Randy,
>>>>>
>>>>> You have two possible choices. It can be configured with 4 x 6v 4Ah
>>>>> batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
>>>>> modified the battery pack already. A couple of mine needed a nibbler
>>> tool
>>>>> to remove enough of the aluminum cover that fits over the tops of
> the
>>>>> batteries. The original cover will short out to the battery tabs
>>> regardless
>>>>> of the battery configuration if this is not done.
>>>>>
>>>>> You can find larger capacity batteries that will give you slightly
> more
>>>>> battery life. I lost a couple sets of mail-order batteries after a
> few
>>>>> extended outages. I would recommend going with locally bought
> batteries
>>>>> instead of the cheaper mail order. My local Batteries Plus will
>>> typically
>>>>> have some warranty if I remember correctly. Moving forward I will
> only
>>> use
>>>>> 2 12v batteries and pre-charge them on a battery charger to equalize
>>> them
>>>>> before putting them in the 732A. I think the cheap batteries did not
>>>>> discharge equally, and would not recover when power was applied.
>>>>>
>>>>> Inspect the back plane for damaged traces and look at the
> capacitors. I
>>> had
>>>>> a few that looked questionable. So far, I have replaced all the big
>>> caps
>>> on
>>>>> the pre-regulator and regulator boards. My feeling is that once
> these
>>> go
>>>>> online, they should run as long as possible between repairs.
>>>>>
>>>>> The battery charger circuit may need adjusting. I tweaked mine and
> it
>>>>> seemed to work fine.
>>>>>
>>>>> Todd
>>>>>
>>>>>
>>>>> On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
>>> randyevans2688@gmail.com <javascript:;>>
>>>>> wrote:
>>>>>
>>>>>> I received my Fluke 732A today. Just powered it up but it needs
> new
>>>>>> batteries. Any suggestions for sources (I haven't opened up the
> unit
>>>>> yet -
>>>>>> I want to make sure it works before doing that). Also received
> the
>>>>>> ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> data
>>>>> dumper
>>>>>> program to get the CAL data from my 3458A. Should be a busy
> weekend.
>>>>>>
>>>>>> Randy
>>>>>> _______________________________________________
>>>>>> volt-nuts mailing list -- volt-nuts@febo.com <javascript:;>
>>>>>> To unsubscribe, go to
>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>>>> and follow the instructions there.
>>>>> _______________________________________________
>>>>> volt-nuts mailing list -- volt-nuts@febo.com <javascript:;>
>>>>> To unsubscribe, go to
>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>>> and follow the instructions there.
>>>> _______________________________________________
>>>> volt-nuts mailing list -- volt-nuts@febo.com <javascript:;>
>>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>> and follow the instructions there.
>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com <javascript:;>
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com <javascript:;>
>> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com <javascript:;>
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
RE
Randy Evans
Mon, Aug 25, 2014 12:01 AM
Bill,
Thanks for the MATH overview. That could be very helpful. I definately
will give it a try.
Randy
On Sun, Aug 24, 2014 at 2:44 PM, Bill Gold wpgold3637@att.net wrote:
Todd:
I check the batteries every month per the 732A manual. Usually I keep
a
few used cells from previous packs around that seem to be ok after checking
them for capacity just like checking the battery pack. I can usually spot
cells going bad with that method. If I don't have a spare set then I just
substitute a used cell until I receive a new set.
I agree with you that the cheap ones are crap. I have used Power Sonic
a lot with some a few EaglePicher and one set of Enersys. The EaglePicher
are crap so I will never use them. EaglePicher have had a couple out of 8
that have failed very early. I have never tried Panasonic. So I stick
with
the Power Sonic's. I am not sure why the 12 volt ones would be better than
the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3
lead acid cells. If one of the 6 cells in the 12 volt battery go bad then
the whole battery has to be changed. So you better buy good quality 12
volt
batteries.
I just can't seem to source batteries locally easily at a reasonable
price. So I order them from Allied usually.
Bill
----- Original Message -----
From: "Todd Micallef" tmicallef@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 8:13 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But
I was mistaken. The 12V 5ah batteries are the ones that I am using in one
of my 732A. There is more play inside the tray with the 12v batteries by
several mm as compared to the 6v which only has a 2-3mm. The 12v
is not difficult, but it is easier if the battery tray has already been
machined for the 12v batteries. The battery tabs don't line up well with
the existing holes and need to be widened. Two additional holes must also
be added.
Sourcing the batteries locally is more of a convenience than waiting for
the delivery. The el-cheapo 6v batteries are a waste of money and I have
batteries so far to prove it. They died the first time I had a 6hr outage
and would not hold a charge after that.
Todd
On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Bill,
Thanks for the MATH overview. That could be very helpful. I definately
will give it a try.
Randy
On Sun, Aug 24, 2014 at 2:44 PM, Bill Gold <wpgold3637@att.net> wrote:
> Todd:
>
> I check the batteries every month per the 732A manual. Usually I keep
> a
> few used cells from previous packs around that seem to be ok after checking
> them for capacity just like checking the battery pack. I can usually spot
> cells going bad with that method. If I don't have a spare set then I just
> substitute a used cell until I receive a new set.
>
> I agree with you that the cheap ones are crap. I have used Power Sonic
> a lot with some a few EaglePicher and one set of Enersys. The EaglePicher
> are crap so I will never use them. EaglePicher have had a couple out of 8
> that have failed very early. I have never tried Panasonic. So I stick
> with
> the Power Sonic's. I am not sure why the 12 volt ones would be better than
> the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3
> lead acid cells. If one of the 6 cells in the 12 volt battery go bad then
> the whole battery has to be changed. So you better buy good quality 12
> volt
> batteries.
>
> I just can't seem to source batteries locally easily at a reasonable
> price. So I order them from Allied usually.
>
> Bill
>
> ----- Original Message -----
> From: "Todd Micallef" <tmicallef@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Sunday, August 24, 2014 8:13 AM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Bill,
> >
> > I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But
> yes,
> > I was mistaken. The 12V 5ah batteries are the ones that I am using in one
> > of my 732A. There is more play inside the tray with the 12v batteries by
> > several mm as compared to the 6v which only has a 2-3mm. The 12v
> conversion
> > is not difficult, but it is easier if the battery tray has already been
> > machined for the 12v batteries. The battery tabs don't line up well with
> > the existing holes and need to be widened. Two additional holes must also
> > be added.
> >
> > Sourcing the batteries locally is more of a convenience than waiting for
> > the delivery. The el-cheapo 6v batteries are a waste of money and I have
> 8
> > batteries so far to prove it. They died the first time I had a 6hr outage
> > and would not hold a charge after that.
> >
> > Todd
> >
> >
> >
> >
> > On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold <wpgold3637@att.net> wrote:
> >
> > > Randy:
> > >
> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> > > others
> > > that are in this size and package. Order from one of the usual
> electronics
> > > distributors like Allied, Mouser, Digikey. This is a very common
> battery
> > > as
> > > it is used in a lot of "EXIT" signs so they are lighted when the power
> goes
> > > out. I don't see how the 12V 7AH will fit as they are too large. I
> guess
> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
> AH
> > > but the terminals are in the wrong place so you will have to "nibble"
> out
> > > the aluminum plate that holds them in the 732A battery pack. You have
> to
> > > be
> > > careful if you use the 12v 5AH as you will have 4 extra battery
> connection
> > > leads to deal with and connect correctly. I would stick with the 6V
> 4AH.
> > > New batteries will last around 12 to 14 hours before the "CAL" light
> goes
> > > out when AC power is not applied. So shipping to Cal Lab can be a
> problem
> > > if it is a distance away, or you have to use a shipper like UPS or
> FEDEX
> > > and
> > > you ship the night before and then use their "Morning delivery" and the
> Cal
> > > Lab is expecting your 732A. Same on the way back to you. Of course
> you
> > > could always strap another battery on the 732A and hook it up to the
> "ext
> > > power" plug to last longer. I have seen it done. The issue is to get
> the
> > > Cal Lab to charge the extra battery before they ship the 732A back to
> you.
> > >
> > > When you remove and work on the battery pack always have the AC
> power
> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> supply
> > > is working.
> > >
> > > The "CAL" light is to indicate that power has not been lost to the
> > > Reference Amp or other associated circuits. When the raw supply
> (battery)
> > > voltage drops below about 21 volts the "CAL" light will go out. Below
> that
> > > voltage the heater circuits will not work correctly and the 18.6 volt
> > > regulated supply will not regulate. The requirement is that the
> Reference
> > > Amp be kept "alive" at all times to maintain the output voltage that
> was
> > > measured at the time of the most recent Calibration or Certification.
> When
> > > the semiconductor junctions are unbiased and cool off when power is
> lost,
> > > and then power is restored the result will be a different 10 volts than
> > > before the power failure. My experience is that after all of the years
> > > that
> > > these units have been powered up, this won't happen and when power is
> lost
> > > and then restored, even months later, the 732A will come back to almost
> > > exactly the same 10 volts as when they lost power, usually with in 0.2
> PPM
> > > after 24 hours of "warm up".
> > >
> > > What type of hookup leads are you using when measuring the 1 volt
> > > output? What is the PLC set to? I always use 100 PLC to measure this.
> If
> > > you don't have "low thermal" connection leads you can experience uV
> changes
> > > for a minute or more after plugging in the leads due to the "thermals"
> > > generated because of the difference in temperature between the banana
> jacks
> > > on the 732A and the banana plugs of the leads. I have found that even
> just
> > > plugging in the lead will generate a thermal difference because of
> > > difference of temps and some heating due to the physical act of just
> > > inserting the plug because of friction between the jack and plug (my
> theory
> > > at any rate). You have to allow at least a minute or more before being
> > > able
> > > to make a measurement after plugging in the leads. I just measured the
> > > variation of the 1 volt output of my 732A and using my 3458A and I got
> a
> > > total difference of 0.159 uV over 40 measurements using 100 PLC on the
> 1
> > > volt range of the 3458A. Using the MATH function and all of the data
> you
> > > can collect. That was after waiting for several minutes after plugging
> in
> > > the leads.
> > >
> > > I hope all of this helps.
> > >
> > > Bill
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > > Sent: Friday, August 22, 2014 9:03 PM
> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >
> > >
> > > > Todd,
> > > >
> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> that
> I
> > > > keep topped off and they have very low current draw (~2 to 3 mA at
> 13.5
> > > > VDC) when charged and at their float voltage, so I am pretty sure
> they
> > > are
> > > > in good condition. I will look at getting those in the units after I
> > > > ascertain the condition of the 732.
> > > >
> > > > So now I have a what appears to be a functioning 3458A and a 732A but
> > > they
> > > > slightly disagree. I am like the man with two watches that disagree
> on
> > > the
> > > > time - which is correct? For the moment, i am only concerned with
> > > > stability. The need for absolute accuracy will come later.
> > > >
> > > > Randy
> > > >
> > > >
> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> > > wrote:
> > > >
> > > > > Randy,
> > > > >
> > > > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> has
> > > > > modified the battery pack already. A couple of mine needed a
> nibbler
> > > tool
> > > > > to remove enough of the aluminum cover that fits over the tops of
> the
> > > > > batteries. The original cover will short out to the battery tabs
> > > regardless
> > > > > of the battery configuration if this is not done.
> > > > >
> > > > > You can find larger capacity batteries that will give you slightly
> more
> > > > > battery life. I lost a couple sets of mail-order batteries after a
> few
> > > > > extended outages. I would recommend going with locally bought
> batteries
> > > > > instead of the cheaper mail order. My local Batteries Plus will
> > > typically
> > > > > have some warranty if I remember correctly. Moving forward I will
> only
> > > use
> > > > > 2 12v batteries and pre-charge them on a battery charger to
> equalize
> > > them
> > > > > before putting them in the 732A. I think the cheap batteries did
> not
> > > > > discharge equally, and would not recover when power was applied.
> > > > >
> > > > > Inspect the back plane for damaged traces and look at the
> capacitors. I
> > > had
> > > > > a few that looked questionable. So far, I have replaced all the big
> > > caps
> > > on
> > > > > the pre-regulator and regulator boards. My feeling is that once
> these
> > > go
> > > > > online, they should run as long as possible between repairs.
> > > > >
> > > > > The battery charger circuit may need adjusting. I tweaked mine and
> it
> > > > > seemed to work fine.
> > > > >
> > > > > Todd
> > > > >
> > > > >
> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > randyevans2688@gmail.com>
> > > > > wrote:
> > > > >
> > > > > > I received my Fluke 732A today. Just powered it up but it needs
> new
> > > > > > batteries. Any suggestions for sources (I haven't opened up the
> unit
> > > > > yet -
> > > > > > I want to make sure it works before doing that). Also received
> the
> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> data
> > > > > dumper
> > > > > > program to get the CAL data from my 3458A. Should be a busy
> weekend.
> > > > > >
> > > > > > Randy
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > To unsubscribe, go to
> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
RE
Randy Evans
Mon, Aug 25, 2014 12:31 AM
Bill,
I am trying to figure out the MATH function without much success. The 3458
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts. This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Bill,
I am trying to figure out the MATH function without much success. The 3458
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> Randy:
>
> The MATH function is accessible from the keypad. I don't have an IEEE
> interface right now that works. You can also program the numeric keypad
> keys to have preprogrammed functions. DEFKEY
>
> I have made my own "low thermal" measurement leads from Pomona #4892
> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
> to do custom cables with 16 ga. bare copper wire that I will twist and then
> put a braided shield over it. I simply cannot find what I want so I will
> build my own cable. I have done something like this before and it worked
> fine. When I get a "round toit".
>
> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> used
> in the past to verify my homemade "low thermal" leads as described above.
> Frankly I cannot see any difference between using the 1756 cables and my
> homemade cables once I give them a few minutes for the thermals to go away.
> As far as I can tell and measure the differences, if any, are below 0.1 ppm
> at 10 volts.
>
> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
> the instability of the readings it is hard to determine which is causing
> the
> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
> set
> the MATH to "Statistics" (store high reading/low reading/ and mean of the
> readings) in the registers, the number of readings to "40", the trigger to
> "hold" (which keeps the meter from triggering until I press "ENTER" and
> then
> trigger the sequence of 40 readings when I push the "ENTER" button. You
> can
> do all of this manually from the keypads but since I use this sequence a
> lot
> I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
> Once those 40 readings are finished then you can access the various MATH
> statistic registers, using the menu, by entering "MATH" and then a 2 for
> low, a 4 for mean, and 13 for high. Of course you could do all of this
> through the IEEE also. The 3458A has a very rich set of measurement
> commands. I am still learning all of them. It depends upon what I am
> trying to accomplish.
>
> Since the 1.018 and 1.0 volt outputs are passive and derived from
> resistive dividers from the 10 volt, I don't see how they could contribute
> to the varying readings you are measuring. I think I would put a short on
> the input of the 3458A and manually set the range to 1 volt and then
> observe
> the variations that way without the 732A involved. When I do this I see a
> variation from low reading to high reading of 0.125 uVolts and then another
> 40 I get 0.155 uVolts. This is without the GUARD connected to the low side
> of the measurment terminals, GUARD connected doesn't seem to affect the
> readings. So that is the base noise of the 3458A without the 732A,
> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
> a variation of 0.159 uVolts using the same 40 reading method above. I
> would
> use this to determine where your problem might exist. Just having the
> meter
> input shorted will point you in the right direction. Meter, cables or
> 732A.
>
> Sorry for the long dissertation. Friends get mad at me for being so
> detailed sometimes.
>
> Bill
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Sunday, August 24, 2014 7:22 AM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Bill,
> >
> > I have convinced myself that the problem I an seeing is due to thermals.
> > If I move the cables (with gold-plated banana plugs) using a small towel
> > rather than letting my hand touch the plugs, it is much more stable. If
> I
> > then hold the banana plug with my hand after the reading has stabilized,
> > the reading drifts rapidly upward. I am trying to check the stability of
> > the reading but I haven't figured out the MATH function yet. I assume
> this
> > is a programmed function using GPIB only?
> >
> > The stability I am seeing by manually recording readings (using NLPC of
> 100
> > and 1000) is much greater than what you are measuring on your system.
> Not
> > sure how to ascertain if it's the 3458A or the 732. The value of the
> > readings are very different between the two - the 3458 reads about 50 uV
> > high on the 10 V output and about 12 uV low on the 1V output. Rather
> large
> > differences (this is after an ACAL). I need to find some better cables
> to
> > make sure the errors are not due to thermals again.
> >
> > Randy
> >
> >
> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
> >
> > > Randy:
> > >
> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> > > others
> > > that are in this size and package. Order from one of the usual
> electronics
> > > distributors like Allied, Mouser, Digikey. This is a very common
> battery
> > > as
> > > it is used in a lot of "EXIT" signs so they are lighted when the power
> goes
> > > out. I don't see how the 12V 7AH will fit as they are too large. I
> guess
> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
> AH
> > > but the terminals are in the wrong place so you will have to "nibble"
> out
> > > the aluminum plate that holds them in the 732A battery pack. You have
> to
> > > be
> > > careful if you use the 12v 5AH as you will have 4 extra battery
> connection
> > > leads to deal with and connect correctly. I would stick with the 6V
> 4AH.
> > > New batteries will last around 12 to 14 hours before the "CAL" light
> goes
> > > out when AC power is not applied. So shipping to Cal Lab can be a
> problem
> > > if it is a distance away, or you have to use a shipper like UPS or
> FEDEX
> > > and
> > > you ship the night before and then use their "Morning delivery" and the
> Cal
> > > Lab is expecting your 732A. Same on the way back to you. Of course
> you
> > > could always strap another battery on the 732A and hook it up to the
> "ext
> > > power" plug to last longer. I have seen it done. The issue is to get
> the
> > > Cal Lab to charge the extra battery before they ship the 732A back to
> you.
> > >
> > > When you remove and work on the battery pack always have the AC
> power
> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> supply
> > > is working.
> > >
> > > The "CAL" light is to indicate that power has not been lost to the
> > > Reference Amp or other associated circuits. When the raw supply
> (battery)
> > > voltage drops below about 21 volts the "CAL" light will go out. Below
> that
> > > voltage the heater circuits will not work correctly and the 18.6 volt
> > > regulated supply will not regulate. The requirement is that the
> Reference
> > > Amp be kept "alive" at all times to maintain the output voltage that
> was
> > > measured at the time of the most recent Calibration or Certification.
> When
> > > the semiconductor junctions are unbiased and cool off when power is
> lost,
> > > and then power is restored the result will be a different 10 volts than
> > > before the power failure. My experience is that after all of the years
> > > that
> > > these units have been powered up, this won't happen and when power is
> lost
> > > and then restored, even months later, the 732A will come back to almost
> > > exactly the same 10 volts as when they lost power, usually with in 0.2
> PPM
> > > after 24 hours of "warm up".
> > >
> > > What type of hookup leads are you using when measuring the 1 volt
> > > output? What is the PLC set to? I always use 100 PLC to measure this.
> If
> > > you don't have "low thermal" connection leads you can experience uV
> changes
> > > for a minute or more after plugging in the leads due to the "thermals"
> > > generated because of the difference in temperature between the banana
> jacks
> > > on the 732A and the banana plugs of the leads. I have found that even
> just
> > > plugging in the lead will generate a thermal difference because of
> > > difference of temps and some heating due to the physical act of just
> > > inserting the plug because of friction between the jack and plug (my
> theory
> > > at any rate). You have to allow at least a minute or more before being
> > > able
> > > to make a measurement after plugging in the leads. I just measured the
> > > variation of the 1 volt output of my 732A and using my 3458A and I got
> a
> > > total difference of 0.159 uV over 40 measurements using 100 PLC on the
> 1
> > > volt range of the 3458A. Using the MATH function and all of the data
> you
> > > can collect. That was after waiting for several minutes after plugging
> in
> > > the leads.
> > >
> > > I hope all of this helps.
> > >
> > > Bill
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > > Sent: Friday, August 22, 2014 9:03 PM
> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >
> > >
> > > > Todd,
> > > >
> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> that
> I
> > > > keep topped off and they have very low current draw (~2 to 3 mA at
> 13.5
> > > > VDC) when charged and at their float voltage, so I am pretty sure
> they
> > > are
> > > > in good condition. I will look at getting those in the units after I
> > > > ascertain the condition of the 732.
> > > >
> > > > So now I have a what appears to be a functioning 3458A and a 732A but
> > > they
> > > > slightly disagree. I am like the man with two watches that disagree
> on
> > > the
> > > > time - which is correct? For the moment, i am only concerned with
> > > > stability. The need for absolute accuracy will come later.
> > > >
> > > > Randy
> > > >
> > > >
> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> > > wrote:
> > > >
> > > > > Randy,
> > > > >
> > > > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> has
> > > > > modified the battery pack already. A couple of mine needed a
> nibbler
> > > tool
> > > > > to remove enough of the aluminum cover that fits over the tops of
> the
> > > > > batteries. The original cover will short out to the battery tabs
> > > regardless
> > > > > of the battery configuration if this is not done.
> > > > >
> > > > > You can find larger capacity batteries that will give you slightly
> more
> > > > > battery life. I lost a couple sets of mail-order batteries after a
> few
> > > > > extended outages. I would recommend going with locally bought
> batteries
> > > > > instead of the cheaper mail order. My local Batteries Plus will
> > > typically
> > > > > have some warranty if I remember correctly. Moving forward I will
> only
> > > use
> > > > > 2 12v batteries and pre-charge them on a battery charger to
> equalize
> > > them
> > > > > before putting them in the 732A. I think the cheap batteries did
> not
> > > > > discharge equally, and would not recover when power was applied.
> > > > >
> > > > > Inspect the back plane for damaged traces and look at the
> capacitors. I
> > > had
> > > > > a few that looked questionable. So far, I have replaced all the big
> > > caps
> > > on
> > > > > the pre-regulator and regulator boards. My feeling is that once
> these
> > > go
> > > > > online, they should run as long as possible between repairs.
> > > > >
> > > > > The battery charger circuit may need adjusting. I tweaked mine and
> it
> > > > > seemed to work fine.
> > > > >
> > > > > Todd
> > > > >
> > > > >
> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > randyevans2688@gmail.com>
> > > > > wrote:
> > > > >
> > > > > > I received my Fluke 732A today. Just powered it up but it needs
> new
> > > > > > batteries. Any suggestions for sources (I haven't opened up the
> unit
> > > > > yet -
> > > > > > I want to make sure it works before doing that). Also received
> the
> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> data
> > > > > dumper
> > > > > > program to get the CAL data from my 3458A. Should be a busy
> weekend.
> > > > > >
> > > > > > Randy
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > To unsubscribe, go to
> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
RE
Randy Evans
Mon, Aug 25, 2014 1:04 AM
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and the
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts. This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and the
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> Randy:
>
> The MATH function is accessible from the keypad. I don't have an IEEE
> interface right now that works. You can also program the numeric keypad
> keys to have preprogrammed functions. DEFKEY
>
> I have made my own "low thermal" measurement leads from Pomona #4892
> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
> to do custom cables with 16 ga. bare copper wire that I will twist and then
> put a braided shield over it. I simply cannot find what I want so I will
> build my own cable. I have done something like this before and it worked
> fine. When I get a "round toit".
>
> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> used
> in the past to verify my homemade "low thermal" leads as described above.
> Frankly I cannot see any difference between using the 1756 cables and my
> homemade cables once I give them a few minutes for the thermals to go away.
> As far as I can tell and measure the differences, if any, are below 0.1 ppm
> at 10 volts.
>
> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
> the instability of the readings it is hard to determine which is causing
> the
> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
> set
> the MATH to "Statistics" (store high reading/low reading/ and mean of the
> readings) in the registers, the number of readings to "40", the trigger to
> "hold" (which keeps the meter from triggering until I press "ENTER" and
> then
> trigger the sequence of 40 readings when I push the "ENTER" button. You
> can
> do all of this manually from the keypads but since I use this sequence a
> lot
> I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
> Once those 40 readings are finished then you can access the various MATH
> statistic registers, using the menu, by entering "MATH" and then a 2 for
> low, a 4 for mean, and 13 for high. Of course you could do all of this
> through the IEEE also. The 3458A has a very rich set of measurement
> commands. I am still learning all of them. It depends upon what I am
> trying to accomplish.
>
> Since the 1.018 and 1.0 volt outputs are passive and derived from
> resistive dividers from the 10 volt, I don't see how they could contribute
> to the varying readings you are measuring. I think I would put a short on
> the input of the 3458A and manually set the range to 1 volt and then
> observe
> the variations that way without the 732A involved. When I do this I see a
> variation from low reading to high reading of 0.125 uVolts and then another
> 40 I get 0.155 uVolts. This is without the GUARD connected to the low side
> of the measurment terminals, GUARD connected doesn't seem to affect the
> readings. So that is the base noise of the 3458A without the 732A,
> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
> a variation of 0.159 uVolts using the same 40 reading method above. I
> would
> use this to determine where your problem might exist. Just having the
> meter
> input shorted will point you in the right direction. Meter, cables or
> 732A.
>
> Sorry for the long dissertation. Friends get mad at me for being so
> detailed sometimes.
>
> Bill
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Sunday, August 24, 2014 7:22 AM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Bill,
> >
> > I have convinced myself that the problem I an seeing is due to thermals.
> > If I move the cables (with gold-plated banana plugs) using a small towel
> > rather than letting my hand touch the plugs, it is much more stable. If
> I
> > then hold the banana plug with my hand after the reading has stabilized,
> > the reading drifts rapidly upward. I am trying to check the stability of
> > the reading but I haven't figured out the MATH function yet. I assume
> this
> > is a programmed function using GPIB only?
> >
> > The stability I am seeing by manually recording readings (using NLPC of
> 100
> > and 1000) is much greater than what you are measuring on your system.
> Not
> > sure how to ascertain if it's the 3458A or the 732. The value of the
> > readings are very different between the two - the 3458 reads about 50 uV
> > high on the 10 V output and about 12 uV low on the 1V output. Rather
> large
> > differences (this is after an ACAL). I need to find some better cables
> to
> > make sure the errors are not due to thermals again.
> >
> > Randy
> >
> >
> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
> >
> > > Randy:
> > >
> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> > > others
> > > that are in this size and package. Order from one of the usual
> electronics
> > > distributors like Allied, Mouser, Digikey. This is a very common
> battery
> > > as
> > > it is used in a lot of "EXIT" signs so they are lighted when the power
> goes
> > > out. I don't see how the 12V 7AH will fit as they are too large. I
> guess
> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
> AH
> > > but the terminals are in the wrong place so you will have to "nibble"
> out
> > > the aluminum plate that holds them in the 732A battery pack. You have
> to
> > > be
> > > careful if you use the 12v 5AH as you will have 4 extra battery
> connection
> > > leads to deal with and connect correctly. I would stick with the 6V
> 4AH.
> > > New batteries will last around 12 to 14 hours before the "CAL" light
> goes
> > > out when AC power is not applied. So shipping to Cal Lab can be a
> problem
> > > if it is a distance away, or you have to use a shipper like UPS or
> FEDEX
> > > and
> > > you ship the night before and then use their "Morning delivery" and the
> Cal
> > > Lab is expecting your 732A. Same on the way back to you. Of course
> you
> > > could always strap another battery on the 732A and hook it up to the
> "ext
> > > power" plug to last longer. I have seen it done. The issue is to get
> the
> > > Cal Lab to charge the extra battery before they ship the 732A back to
> you.
> > >
> > > When you remove and work on the battery pack always have the AC
> power
> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> supply
> > > is working.
> > >
> > > The "CAL" light is to indicate that power has not been lost to the
> > > Reference Amp or other associated circuits. When the raw supply
> (battery)
> > > voltage drops below about 21 volts the "CAL" light will go out. Below
> that
> > > voltage the heater circuits will not work correctly and the 18.6 volt
> > > regulated supply will not regulate. The requirement is that the
> Reference
> > > Amp be kept "alive" at all times to maintain the output voltage that
> was
> > > measured at the time of the most recent Calibration or Certification.
> When
> > > the semiconductor junctions are unbiased and cool off when power is
> lost,
> > > and then power is restored the result will be a different 10 volts than
> > > before the power failure. My experience is that after all of the years
> > > that
> > > these units have been powered up, this won't happen and when power is
> lost
> > > and then restored, even months later, the 732A will come back to almost
> > > exactly the same 10 volts as when they lost power, usually with in 0.2
> PPM
> > > after 24 hours of "warm up".
> > >
> > > What type of hookup leads are you using when measuring the 1 volt
> > > output? What is the PLC set to? I always use 100 PLC to measure this.
> If
> > > you don't have "low thermal" connection leads you can experience uV
> changes
> > > for a minute or more after plugging in the leads due to the "thermals"
> > > generated because of the difference in temperature between the banana
> jacks
> > > on the 732A and the banana plugs of the leads. I have found that even
> just
> > > plugging in the lead will generate a thermal difference because of
> > > difference of temps and some heating due to the physical act of just
> > > inserting the plug because of friction between the jack and plug (my
> theory
> > > at any rate). You have to allow at least a minute or more before being
> > > able
> > > to make a measurement after plugging in the leads. I just measured the
> > > variation of the 1 volt output of my 732A and using my 3458A and I got
> a
> > > total difference of 0.159 uV over 40 measurements using 100 PLC on the
> 1
> > > volt range of the 3458A. Using the MATH function and all of the data
> you
> > > can collect. That was after waiting for several minutes after plugging
> in
> > > the leads.
> > >
> > > I hope all of this helps.
> > >
> > > Bill
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > > Sent: Friday, August 22, 2014 9:03 PM
> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >
> > >
> > > > Todd,
> > > >
> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> that
> I
> > > > keep topped off and they have very low current draw (~2 to 3 mA at
> 13.5
> > > > VDC) when charged and at their float voltage, so I am pretty sure
> they
> > > are
> > > > in good condition. I will look at getting those in the units after I
> > > > ascertain the condition of the 732.
> > > >
> > > > So now I have a what appears to be a functioning 3458A and a 732A but
> > > they
> > > > slightly disagree. I am like the man with two watches that disagree
> on
> > > the
> > > > time - which is correct? For the moment, i am only concerned with
> > > > stability. The need for absolute accuracy will come later.
> > > >
> > > > Randy
> > > >
> > > >
> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
> > > wrote:
> > > >
> > > > > Randy,
> > > > >
> > > > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> has
> > > > > modified the battery pack already. A couple of mine needed a
> nibbler
> > > tool
> > > > > to remove enough of the aluminum cover that fits over the tops of
> the
> > > > > batteries. The original cover will short out to the battery tabs
> > > regardless
> > > > > of the battery configuration if this is not done.
> > > > >
> > > > > You can find larger capacity batteries that will give you slightly
> more
> > > > > battery life. I lost a couple sets of mail-order batteries after a
> few
> > > > > extended outages. I would recommend going with locally bought
> batteries
> > > > > instead of the cheaper mail order. My local Batteries Plus will
> > > typically
> > > > > have some warranty if I remember correctly. Moving forward I will
> only
> > > use
> > > > > 2 12v batteries and pre-charge them on a battery charger to
> equalize
> > > them
> > > > > before putting them in the 732A. I think the cheap batteries did
> not
> > > > > discharge equally, and would not recover when power was applied.
> > > > >
> > > > > Inspect the back plane for damaged traces and look at the
> capacitors. I
> > > had
> > > > > a few that looked questionable. So far, I have replaced all the big
> > > caps
> > > on
> > > > > the pre-regulator and regulator boards. My feeling is that once
> these
> > > go
> > > > > online, they should run as long as possible between repairs.
> > > > >
> > > > > The battery charger circuit may need adjusting. I tweaked mine and
> it
> > > > > seemed to work fine.
> > > > >
> > > > > Todd
> > > > >
> > > > >
> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > randyevans2688@gmail.com>
> > > > > wrote:
> > > > >
> > > > > > I received my Fluke 732A today. Just powered it up but it needs
> new
> > > > > > batteries. Any suggestions for sources (I haven't opened up the
> unit
> > > > > yet -
> > > > > > I want to make sure it works before doing that). Also received
> the
> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> data
> > > > > dumper
> > > > > > program to get the CAL data from my 3458A. Should be a busy
> weekend.
> > > > > >
> > > > > > Randy
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > To unsubscribe, go to
> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
CB
Charles Black
Mon, Aug 25, 2014 2:18 AM
Hi Randy,
You can use the NPLC command to change the A/D converter's'integration
time. Set it to 1000 if you want full resolution. It sets the hp3458a's NMR.
Charlie
On 8/24/2014 6:04 PM, Randy Evans wrote:
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and the
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts. This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable. If
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability of
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and the
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts than
before the power failure. My experience is that after all of the years
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure this.
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads. I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got
total difference of 0.159 uV over 40 measurements using 100 PLC on the
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after plugging
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after I
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A but
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the big
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Hi Randy,
You can use the NPLC command to change the A/D converter's'integration
time. Set it to 1000 if you want full resolution. It sets the hp3458a's NMR.
Charlie
On 8/24/2014 6:04 PM, Randy Evans wrote:
> Bill,
>
> I am trying to figure out the MATH function without much success. I input
> the sequence you said (I looked up the instructions to understand what you
> did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
> it shows up on the display when I input BLUE F1. I hit ENTER and it takes
> the 40 measurements and the MATH symbol shows on the display during the
> measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> get a MATH ERR symbol on the display. I tried it a couple of times and the
> same result so I am doing something wrong. Is there a better source for
> explaining how to do front panel masurements than the User Guide, which
> seems oriented at programming automatic rather than manual measurements.
>
> Randy
>
>
>
> On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
>
>> Randy:
>>
>> The MATH function is accessible from the keypad. I don't have an IEEE
>> interface right now that works. You can also program the numeric keypad
>> keys to have preprogrammed functions. DEFKEY
>>
>> I have made my own "low thermal" measurement leads from Pomona #4892
>> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
>> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
>> to do custom cables with 16 ga. bare copper wire that I will twist and then
>> put a braided shield over it. I simply cannot find what I want so I will
>> build my own cable. I have done something like this before and it worked
>> fine. When I get a "round toit".
>>
>> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
>> used
>> in the past to verify my homemade "low thermal" leads as described above.
>> Frankly I cannot see any difference between using the 1756 cables and my
>> homemade cables once I give them a few minutes for the thermals to go away.
>> As far as I can tell and measure the differences, if any, are below 0.1 ppm
>> at 10 volts.
>>
>> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
>> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
>> the instability of the readings it is hard to determine which is causing
>> the
>> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
>> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
>> set
>> the MATH to "Statistics" (store high reading/low reading/ and mean of the
>> readings) in the registers, the number of readings to "40", the trigger to
>> "hold" (which keeps the meter from triggering until I press "ENTER" and
>> then
>> trigger the sequence of 40 readings when I push the "ENTER" button. You
>> can
>> do all of this manually from the keypads but since I use this sequence a
>> lot
>> I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
>> Once those 40 readings are finished then you can access the various MATH
>> statistic registers, using the menu, by entering "MATH" and then a 2 for
>> low, a 4 for mean, and 13 for high. Of course you could do all of this
>> through the IEEE also. The 3458A has a very rich set of measurement
>> commands. I am still learning all of them. It depends upon what I am
>> trying to accomplish.
>>
>> Since the 1.018 and 1.0 volt outputs are passive and derived from
>> resistive dividers from the 10 volt, I don't see how they could contribute
>> to the varying readings you are measuring. I think I would put a short on
>> the input of the 3458A and manually set the range to 1 volt and then
>> observe
>> the variations that way without the 732A involved. When I do this I see a
>> variation from low reading to high reading of 0.125 uVolts and then another
>> 40 I get 0.155 uVolts. This is without the GUARD connected to the low side
>> of the measurment terminals, GUARD connected doesn't seem to affect the
>> readings. So that is the base noise of the 3458A without the 732A,
>> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
>> a variation of 0.159 uVolts using the same 40 reading method above. I
>> would
>> use this to determine where your problem might exist. Just having the
>> meter
>> input shorted will point you in the right direction. Meter, cables or
>> 732A.
>>
>> Sorry for the long dissertation. Friends get mad at me for being so
>> detailed sometimes.
>>
>> Bill
>>
>> ----- Original Message -----
>> From: "Randy Evans" <randyevans2688@gmail.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Sent: Sunday, August 24, 2014 7:22 AM
>> Subject: Re: [volt-nuts] 732A and Prologix received
>>
>>
>>> Bill,
>>>
>>> I have convinced myself that the problem I an seeing is due to thermals.
>>> If I move the cables (with gold-plated banana plugs) using a small towel
>>> rather than letting my hand touch the plugs, it is much more stable. If
>> I
>>> then hold the banana plug with my hand after the reading has stabilized,
>>> the reading drifts rapidly upward. I am trying to check the stability of
>>> the reading but I haven't figured out the MATH function yet. I assume
>> this
>>> is a programmed function using GPIB only?
>>>
>>> The stability I am seeing by manually recording readings (using NLPC of
>> 100
>>> and 1000) is much greater than what you are measuring on your system.
>> Not
>>> sure how to ascertain if it's the 3458A or the 732. The value of the
>>> readings are very different between the two - the 3458 reads about 50 uV
>>> high on the 10 V output and about 12 uV low on the 1V output. Rather
>> large
>>> differences (this is after an ACAL). I need to find some better cables
>> to
>>> make sure the errors are not due to thermals again.
>>>
>>> Randy
>>>
>>>
>>> On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
>>>
>>>> Randy:
>>>>
>>>> I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
>>>> perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
>>>> others
>>>> that are in this size and package. Order from one of the usual
>> electronics
>>>> distributors like Allied, Mouser, Digikey. This is a very common
>> battery
>>>> as
>>>> it is used in a lot of "EXIT" signs so they are lighted when the power
>> goes
>>>> out. I don't see how the 12V 7AH will fit as they are too large. I
>> guess
>>>> you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
>> AH
>>>> but the terminals are in the wrong place so you will have to "nibble"
>> out
>>>> the aluminum plate that holds them in the 732A battery pack. You have
>> to
>>>> be
>>>> careful if you use the 12v 5AH as you will have 4 extra battery
>> connection
>>>> leads to deal with and connect correctly. I would stick with the 6V
>> 4AH.
>>>> New batteries will last around 12 to 14 hours before the "CAL" light
>> goes
>>>> out when AC power is not applied. So shipping to Cal Lab can be a
>> problem
>>>> if it is a distance away, or you have to use a shipper like UPS or
>> FEDEX
>>>> and
>>>> you ship the night before and then use their "Morning delivery" and the
>> Cal
>>>> Lab is expecting your 732A. Same on the way back to you. Of course
>> you
>>>> could always strap another battery on the 732A and hook it up to the
>> "ext
>>>> power" plug to last longer. I have seen it done. The issue is to get
>> the
>>>> Cal Lab to charge the extra battery before they ship the 732A back to
>> you.
>>>> When you remove and work on the battery pack always have the AC
>> power
>>>> plugged in. The "CAL" led will stay on because the 18.6 v regulated
>> supply
>>>> is working.
>>>>
>>>> The "CAL" light is to indicate that power has not been lost to the
>>>> Reference Amp or other associated circuits. When the raw supply
>> (battery)
>>>> voltage drops below about 21 volts the "CAL" light will go out. Below
>> that
>>>> voltage the heater circuits will not work correctly and the 18.6 volt
>>>> regulated supply will not regulate. The requirement is that the
>> Reference
>>>> Amp be kept "alive" at all times to maintain the output voltage that
>> was
>>>> measured at the time of the most recent Calibration or Certification.
>> When
>>>> the semiconductor junctions are unbiased and cool off when power is
>> lost,
>>>> and then power is restored the result will be a different 10 volts than
>>>> before the power failure. My experience is that after all of the years
>>>> that
>>>> these units have been powered up, this won't happen and when power is
>> lost
>>>> and then restored, even months later, the 732A will come back to almost
>>>> exactly the same 10 volts as when they lost power, usually with in 0.2
>> PPM
>>>> after 24 hours of "warm up".
>>>>
>>>> What type of hookup leads are you using when measuring the 1 volt
>>>> output? What is the PLC set to? I always use 100 PLC to measure this.
>> If
>>>> you don't have "low thermal" connection leads you can experience uV
>> changes
>>>> for a minute or more after plugging in the leads due to the "thermals"
>>>> generated because of the difference in temperature between the banana
>> jacks
>>>> on the 732A and the banana plugs of the leads. I have found that even
>> just
>>>> plugging in the lead will generate a thermal difference because of
>>>> difference of temps and some heating due to the physical act of just
>>>> inserting the plug because of friction between the jack and plug (my
>> theory
>>>> at any rate). You have to allow at least a minute or more before being
>>>> able
>>>> to make a measurement after plugging in the leads. I just measured the
>>>> variation of the 1 volt output of my 732A and using my 3458A and I got
>> a
>>>> total difference of 0.159 uV over 40 measurements using 100 PLC on the
>> 1
>>>> volt range of the 3458A. Using the MATH function and all of the data
>> you
>>>> can collect. That was after waiting for several minutes after plugging
>> in
>>>> the leads.
>>>>
>>>> I hope all of this helps.
>>>>
>>>> Bill
>>>>
>>>>
>>>>
>>>> ----- Original Message -----
>>>> From: "Randy Evans" <randyevans2688@gmail.com>
>>>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>>>> Sent: Friday, August 22, 2014 9:03 PM
>>>> Subject: Re: [volt-nuts] 732A and Prologix received
>>>>
>>>>
>>>>> Todd,
>>>>>
>>>>> Thanks for the info. I have several Panasonic 12V 7 AH batteries
>> that
>> I
>>>>> keep topped off and they have very low current draw (~2 to 3 mA at
>> 13.5
>>>>> VDC) when charged and at their float voltage, so I am pretty sure
>> they
>>>> are
>>>>> in good condition. I will look at getting those in the units after I
>>>>> ascertain the condition of the 732.
>>>>>
>>>>> So now I have a what appears to be a functioning 3458A and a 732A but
>>>> they
>>>>> slightly disagree. I am like the man with two watches that disagree
>> on
>>>> the
>>>>> time - which is correct? For the moment, i am only concerned with
>>>>> stability. The need for absolute accuracy will come later.
>>>>>
>>>>> Randy
>>>>>
>>>>>
>>>>> On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com>
>>>> wrote:
>>>>>> Randy,
>>>>>>
>>>>>> You have two possible choices. It can be configured with 4 x 6v 4Ah
>>>>>> batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
>> has
>>>>>> modified the battery pack already. A couple of mine needed a
>> nibbler
>>>> tool
>>>>>> to remove enough of the aluminum cover that fits over the tops of
>> the
>>>>>> batteries. The original cover will short out to the battery tabs
>>>> regardless
>>>>>> of the battery configuration if this is not done.
>>>>>>
>>>>>> You can find larger capacity batteries that will give you slightly
>> more
>>>>>> battery life. I lost a couple sets of mail-order batteries after a
>> few
>>>>>> extended outages. I would recommend going with locally bought
>> batteries
>>>>>> instead of the cheaper mail order. My local Batteries Plus will
>>>> typically
>>>>>> have some warranty if I remember correctly. Moving forward I will
>> only
>>>> use
>>>>>> 2 12v batteries and pre-charge them on a battery charger to
>> equalize
>>>> them
>>>>>> before putting them in the 732A. I think the cheap batteries did
>> not
>>>>>> discharge equally, and would not recover when power was applied.
>>>>>>
>>>>>> Inspect the back plane for damaged traces and look at the
>> capacitors. I
>>>> had
>>>>>> a few that looked questionable. So far, I have replaced all the big
>>>> caps
>>>> on
>>>>>> the pre-regulator and regulator boards. My feeling is that once
>> these
>>>> go
>>>>>> online, they should run as long as possible between repairs.
>>>>>>
>>>>>> The battery charger circuit may need adjusting. I tweaked mine and
>> it
>>>>>> seemed to work fine.
>>>>>>
>>>>>> Todd
>>>>>>
>>>>>>
>>>>>> On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
>>>> randyevans2688@gmail.com>
>>>>>> wrote:
>>>>>>
>>>>>>> I received my Fluke 732A today. Just powered it up but it needs
>> new
>>>>>>> batteries. Any suggestions for sources (I haven't opened up the
>> unit
>>>>>> yet -
>>>>>>> I want to make sure it works before doing that). Also received
>> the
>>>>>>> ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
>> data
>>>>>> dumper
>>>>>>> program to get the CAL data from my 3458A. Should be a busy
>> weekend.
>>>>>>> Randy
>>>>>>> _______________________________________________
>>>>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>>>>> To unsubscribe, go to
>>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>>>>> and follow the instructions there.
>>>>>>>
>>>>>> _______________________________________________
>>>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>>>> To unsubscribe, go to
>>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>>>> and follow the instructions there.
>>>>>>
>>>>> _______________________________________________
>>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>>> and follow the instructions there.
>>>> _______________________________________________
>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>> and follow the instructions there.
>>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
>
OE
Orin Eman
Mon, Aug 25, 2014 4:33 AM
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs. Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these
as an alternative?
Orin.
On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com>
wrote:
> Randy & all,
>
> You have correctly concluded that some (maybe not all) of your measurement
> problem is thermal EMF being added or subtracted in series within your
> measurement interconnect. This thermal EMF is generated at the junction of
> dissimilar metals when accompanied with thermal gradients between the test
> lead and device terminals. You have to eliminate both the dissimilarity of
> the metal junctions as well as minimize the thermal differences. The
> terminals of the 3458A as well as the 732A are Beryllium Copper so you want
> to use the same test lead terminals. Forget the typical Tin plated lugs or
> even Gold plated as both are not Beryllium Copper and constitute dissimilar
> metals. The best solution (as usually the most expensive) is to use a set
> of
> Fluke 5440A-7005 (48") cables. I also have just as good results using the
> much more flexible Pomona 11174A (lugs end always stay connected to the
> 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
> has the added Guard built in but be sure to also use a Guard lead with the
> Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
> cables is usually not a good idea because the lead wire to terminal also
> constitutes just as critical of junction. The above cables use Tellurium
> Copper wire which is usually hard to find and hard to crimp properly and
> NEVER solder.
>
11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs. Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these
as an alternative?
Orin.
A
acbern@gmx.de
Mon, Aug 25, 2014 10:36 AM
I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.
> Gesendet: Montag, 25. August 2014 um 06:33 Uhr
> Von: "Orin Eman" <orin.eman@gmail.com>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com>
> wrote:
>
> > Randy & all,
> >
> > You have correctly concluded that some (maybe not all) of your measurement
> > problem is thermal EMF being added or subtracted in series within your
> > measurement interconnect. This thermal EMF is generated at the junction of
> > dissimilar metals when accompanied with thermal gradients between the test
> > lead and device terminals. You have to eliminate both the dissimilarity of
> > the metal junctions as well as minimize the thermal differences. The
> > terminals of the 3458A as well as the 732A are Beryllium Copper so you want
> > to use the same test lead terminals. Forget the typical Tin plated lugs or
> > even Gold plated as both are not Beryllium Copper and constitute dissimilar
> > metals. The best solution (as usually the most expensive) is to use a set
> > of
> > Fluke 5440A-7005 (48") cables. I also have just as good results using the
> > much more flexible Pomona 11174A (lugs end always stay connected to the
> > 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
> > has the added Guard built in but be sure to also use a Guard lead with the
> > Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
> > cables is usually not a good idea because the lead wire to terminal also
> > constitutes just as critical of junction. The above cables use Tellurium
> > Copper wire which is usually hard to find and hard to crimp properly and
> > NEVER solder.
> >
>
>
> 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade
> to banana cables are available (5295-36 at Mouser et al) and claim that
> they are designed to minimize thermal EMFs. Datasheet is here:
> http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these
> as an alternative?
>
> Orin.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
BG
Bill Gold
Mon, Aug 25, 2014 1:40 PM
Randy:
Sorry my fault. You have to use the "RMATH" command to get the various
values stored in the registers. See the "RMATH" command in the User's Guide
for a list of what registers you can read.
I sure haven't found any other guides other than the 4 manuals. User's
Guide, Quick Reference Guide, Calibration Manual, and Assembly Level Repair.
It is just a matter of reading the guides and trying to remember what
commands are available. It took me a lot of time to figure out what
commands I use now. I am sure I am missing other commands that might be
useful.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 6:04 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go
As far as I can tell and measure the differences, if any, are below 0.1
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the trigger
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a short
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see
variation from low reading to high reading of 0.125 uVolts and then
40 I get 0.155 uVolts. This is without the GUARD connected to the low
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the stability
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to
Cal Lab to charge the extra battery before they ship the 732A back
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries after
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened up
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Randy:
Sorry my fault. You have to use the "RMATH" command to get the various
values stored in the registers. See the "RMATH" command in the User's Guide
for a list of what registers you can read.
I sure haven't found any other guides other than the 4 manuals. User's
Guide, Quick Reference Guide, Calibration Manual, and Assembly Level Repair.
It is just a matter of reading the guides and trying to remember what
commands are available. It took me a lot of time to figure out what
commands I use now. I am sure I am missing other commands that might be
useful.
Bill
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Sunday, August 24, 2014 6:04 PM
Subject: Re: [volt-nuts] 732A and Prologix received
> Bill,
>
> I am trying to figure out the MATH function without much success. I input
> the sequence you said (I looked up the instructions to understand what you
> did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
and
> it shows up on the display when I input BLUE F1. I hit ENTER and it takes
> the 40 measurements and the MATH symbol shows on the display during the
> measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> get a MATH ERR symbol on the display. I tried it a couple of times and
the
> same result so I am doing something wrong. Is there a better source for
> explaining how to do front panel masurements than the User Guide, which
> seems oriented at programming automatic rather than manual measurements.
>
> Randy
>
>
>
> On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
>
> > Randy:
> >
> > The MATH function is accessible from the keypad. I don't have an
IEEE
> > interface right now that works. You can also program the numeric keypad
> > keys to have preprogrammed functions. DEFKEY
> >
> > I have made my own "low thermal" measurement leads from Pomona #4892
> > banana plugs and Belden #9272 wire. Why 9272, because it was handy at
the
> > time. It is tin plated copper, shielded twisted pair 20 ga. I have
plans
> > to do custom cables with 16 ga. bare copper wire that I will twist and
then
> > put a braided shield over it. I simply cannot find what I want so I
will
> > build my own cable. I have done something like this before and it
worked
> > fine. When I get a "round toit".
> >
> > I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> > used
> > in the past to verify my homemade "low thermal" leads as described
above.
> > Frankly I cannot see any difference between using the 1756 cables and my
> > homemade cables once I give them a few minutes for the thermals to go
away.
> > As far as I can tell and measure the differences, if any, are below 0.1
ppm
> > at 10 volts.
> >
> > Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
all
> > adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
as
> > the instability of the readings it is hard to determine which is causing
> > the
> > problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> > following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
> > set
> > the MATH to "Statistics" (store high reading/low reading/ and mean of
the
> > readings) in the registers, the number of readings to "40", the trigger
to
> > "hold" (which keeps the meter from triggering until I press "ENTER" and
> > then
> > trigger the sequence of 40 readings when I push the "ENTER" button. You
> > can
> > do all of this manually from the keypads but since I use this sequence a
> > lot
> > I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
> > Once those 40 readings are finished then you can access the various MATH
> > statistic registers, using the menu, by entering "MATH" and then a 2 for
> > low, a 4 for mean, and 13 for high. Of course you could do all of this
> > through the IEEE also. The 3458A has a very rich set of measurement
> > commands. I am still learning all of them. It depends upon what I am
> > trying to accomplish.
> >
> > Since the 1.018 and 1.0 volt outputs are passive and derived from
> > resistive dividers from the 10 volt, I don't see how they could
contribute
> > to the varying readings you are measuring. I think I would put a short
on
> > the input of the 3458A and manually set the range to 1 volt and then
> > observe
> > the variations that way without the 732A involved. When I do this I see
a
> > variation from low reading to high reading of 0.125 uVolts and then
another
> > 40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
> > of the measurment terminals, GUARD connected doesn't seem to affect the
> > readings. So that is the base noise of the 3458A without the 732A,
> > somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
> > a variation of 0.159 uVolts using the same 40 reading method above. I
> > would
> > use this to determine where your problem might exist. Just having the
> > meter
> > input shorted will point you in the right direction. Meter, cables or
> > 732A.
> >
> > Sorry for the long dissertation. Friends get mad at me for being so
> > detailed sometimes.
> >
> > Bill
> >
> > ----- Original Message -----
> > From: "Randy Evans" <randyevans2688@gmail.com>
> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Sent: Sunday, August 24, 2014 7:22 AM
> > Subject: Re: [volt-nuts] 732A and Prologix received
> >
> >
> > > Bill,
> > >
> > > I have convinced myself that the problem I an seeing is due to
thermals.
> > > If I move the cables (with gold-plated banana plugs) using a small
towel
> > > rather than letting my hand touch the plugs, it is much more stable.
If
> > I
> > > then hold the banana plug with my hand after the reading has
stabilized,
> > > the reading drifts rapidly upward. I am trying to check the stability
of
> > > the reading but I haven't figured out the MATH function yet. I assume
> > this
> > > is a programmed function using GPIB only?
> > >
> > > The stability I am seeing by manually recording readings (using NLPC
of
> > 100
> > > and 1000) is much greater than what you are measuring on your system.
> > Not
> > > sure how to ascertain if it's the 3458A or the 732. The value of the
> > > readings are very different between the two - the 3458 reads about 50
uV
> > > high on the 10 V output and about 12 uV low on the 1V output. Rather
> > large
> > > differences (this is after an ACAL). I need to find some better
cables
> > to
> > > make sure the errors are not due to thermals again.
> > >
> > > Randy
> > >
> > >
> > > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
wrote:
> > >
> > > > Randy:
> > > >
> > > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
fit
> > > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
and
> > > > others
> > > > that are in this size and package. Order from one of the usual
> > electronics
> > > > distributors like Allied, Mouser, Digikey. This is a very common
> > battery
> > > > as
> > > > it is used in a lot of "EXIT" signs so they are lighted when the
power
> > goes
> > > > out. I don't see how the 12V 7AH will fit as they are too large. I
> > guess
> > > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
volt 4
> > AH
> > > > but the terminals are in the wrong place so you will have to
"nibble"
> > out
> > > > the aluminum plate that holds them in the 732A battery pack. You
have
> > to
> > > > be
> > > > careful if you use the 12v 5AH as you will have 4 extra battery
> > connection
> > > > leads to deal with and connect correctly. I would stick with the 6V
> > 4AH.
> > > > New batteries will last around 12 to 14 hours before the "CAL" light
> > goes
> > > > out when AC power is not applied. So shipping to Cal Lab can be a
> > problem
> > > > if it is a distance away, or you have to use a shipper like UPS or
> > FEDEX
> > > > and
> > > > you ship the night before and then use their "Morning delivery" and
the
> > Cal
> > > > Lab is expecting your 732A. Same on the way back to you. Of course
> > you
> > > > could always strap another battery on the 732A and hook it up to the
> > "ext
> > > > power" plug to last longer. I have seen it done. The issue is to
get
> > the
> > > > Cal Lab to charge the extra battery before they ship the 732A back
to
> > you.
> > > >
> > > > When you remove and work on the battery pack always have the AC
> > power
> > > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> > supply
> > > > is working.
> > > >
> > > > The "CAL" light is to indicate that power has not been lost to
the
> > > > Reference Amp or other associated circuits. When the raw supply
> > (battery)
> > > > voltage drops below about 21 volts the "CAL" light will go out.
Below
> > that
> > > > voltage the heater circuits will not work correctly and the 18.6
volt
> > > > regulated supply will not regulate. The requirement is that the
> > Reference
> > > > Amp be kept "alive" at all times to maintain the output voltage that
> > was
> > > > measured at the time of the most recent Calibration or
Certification.
> > When
> > > > the semiconductor junctions are unbiased and cool off when power is
> > lost,
> > > > and then power is restored the result will be a different 10 volts
than
> > > > before the power failure. My experience is that after all of the
years
> > > > that
> > > > these units have been powered up, this won't happen and when power
is
> > lost
> > > > and then restored, even months later, the 732A will come back to
almost
> > > > exactly the same 10 volts as when they lost power, usually with in
0.2
> > PPM
> > > > after 24 hours of "warm up".
> > > >
> > > > What type of hookup leads are you using when measuring the 1
volt
> > > > output? What is the PLC set to? I always use 100 PLC to measure
this.
> > If
> > > > you don't have "low thermal" connection leads you can experience uV
> > changes
> > > > for a minute or more after plugging in the leads due to the
"thermals"
> > > > generated because of the difference in temperature between the
banana
> > jacks
> > > > on the 732A and the banana plugs of the leads. I have found that
even
> > just
> > > > plugging in the lead will generate a thermal difference because of
> > > > difference of temps and some heating due to the physical act of just
> > > > inserting the plug because of friction between the jack and plug (my
> > theory
> > > > at any rate). You have to allow at least a minute or more before
being
> > > > able
> > > > to make a measurement after plugging in the leads. I just measured
the
> > > > variation of the 1 volt output of my 732A and using my 3458A and I
got
> > a
> > > > total difference of 0.159 uV over 40 measurements using 100 PLC on
the
> > 1
> > > > volt range of the 3458A. Using the MATH function and all of the
data
> > you
> > > > can collect. That was after waiting for several minutes after
plugging
> > in
> > > > the leads.
> > > >
> > > > I hope all of this helps.
> > > >
> > > > Bill
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: "Randy Evans" <randyevans2688@gmail.com>
> > > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > > > Sent: Friday, August 22, 2014 9:03 PM
> > > > Subject: Re: [volt-nuts] 732A and Prologix received
> > > >
> > > >
> > > > > Todd,
> > > > >
> > > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> > that
> > I
> > > > > keep topped off and they have very low current draw (~2 to 3 mA at
> > 13.5
> > > > > VDC) when charged and at their float voltage, so I am pretty sure
> > they
> > > > are
> > > > > in good condition. I will look at getting those in the units
after I
> > > > > ascertain the condition of the 732.
> > > > >
> > > > > So now I have a what appears to be a functioning 3458A and a 732A
but
> > > > they
> > > > > slightly disagree. I am like the man with two watches that
disagree
> > on
> > > > the
> > > > > time - which is correct? For the moment, i am only concerned
with
> > > > > stability. The need for absolute accuracy will come later.
> > > > >
> > > > > Randy
> > > > >
> > > > >
> > > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
<tmicallef@gmail.com>
> > > > wrote:
> > > > >
> > > > > > Randy,
> > > > > >
> > > > > > You have two possible choices. It can be configured with 4 x 6v
4Ah
> > > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> > has
> > > > > > modified the battery pack already. A couple of mine needed a
> > nibbler
> > > > tool
> > > > > > to remove enough of the aluminum cover that fits over the tops
of
> > the
> > > > > > batteries. The original cover will short out to the battery tabs
> > > > regardless
> > > > > > of the battery configuration if this is not done.
> > > > > >
> > > > > > You can find larger capacity batteries that will give you
slightly
> > more
> > > > > > battery life. I lost a couple sets of mail-order batteries after
a
> > few
> > > > > > extended outages. I would recommend going with locally bought
> > batteries
> > > > > > instead of the cheaper mail order. My local Batteries Plus will
> > > > typically
> > > > > > have some warranty if I remember correctly. Moving forward I
will
> > only
> > > > use
> > > > > > 2 12v batteries and pre-charge them on a battery charger to
> > equalize
> > > > them
> > > > > > before putting them in the 732A. I think the cheap batteries did
> > not
> > > > > > discharge equally, and would not recover when power was applied.
> > > > > >
> > > > > > Inspect the back plane for damaged traces and look at the
> > capacitors. I
> > > > had
> > > > > > a few that looked questionable. So far, I have replaced all the
big
> > > > caps
> > > > on
> > > > > > the pre-regulator and regulator boards. My feeling is that once
> > these
> > > > go
> > > > > > online, they should run as long as possible between repairs.
> > > > > >
> > > > > > The battery charger circuit may need adjusting. I tweaked mine
and
> > it
> > > > > > seemed to work fine.
> > > > > >
> > > > > > Todd
> > > > > >
> > > > > >
> > > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > > randyevans2688@gmail.com>
> > > > > > wrote:
> > > > > >
> > > > > > > I received my Fluke 732A today. Just powered it up but it
needs
> > new
> > > > > > > batteries. Any suggestions for sources (I haven't opened up
the
> > unit
> > > > > > yet -
> > > > > > > I want to make sure it works before doing that). Also
received
> > the
> > > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> > data
> > > > > > dumper
> > > > > > > program to get the CAL data from my 3458A. Should be a busy
> > weekend.
> > > > > > >
> > > > > > > Randy
> > > > > > > _______________________________________________
> > > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > > To unsubscribe, go to
> > > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > > and follow the instructions there.
> > > > > > >
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > To unsubscribe, go to
> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
P
pa4tim@gmail.com
Mon, Aug 25, 2014 1:43 PM
Is there a way to link an AC voltage to a DC source for compare. I can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) against standardcells. But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last calibration was 2 years ago.
My idea is in theory simple. It is based on the thermal converters used in RF powermeters. Two resistors, two high resolution temperature meters. AC on the first en DC on the second. If both are the same temperature the AC voltage is the same as the DC voltage. But I'm sure some people here have done this in the past. I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done with lightbubs but I never tryed that) I do not mind if it is slow etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
Verzonden met Windows Mail
Van: Bill Gold
Verzonden: ‎maandag‎ ‎25‎ ‎augustus‎ ‎2014 ‎15‎:‎40
Aan: volt-nuts
Randy:
Sorry my fault. You have to use the "RMATH" command to get the various
values stored in the registers. See the "RMATH" command in the User's Guide
for a list of what registers you can read.
I sure haven't found any other guides other than the 4 manuals. User's
Guide, Quick Reference Guide, Calibration Manual, and Assembly Level Repair.
It is just a matter of reading the guides and trying to remember what
commands are available. It took me a lot of time to figure out what
commands I use now. I am sure I am missing other commands that might be
useful.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 6:04 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go
As far as I can tell and measure the differences, if any, are below 0.1
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the trigger
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a short
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see
variation from low reading to high reading of 0.125 uVolts and then
40 I get 0.155 uVolts. This is without the GUARD connected to the low
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the stability
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to
Cal Lab to charge the extra battery before they ship the 732A back
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries after
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened up
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Is there a way to link an AC voltage to a DC source for compare. I can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) against standardcells. But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last calibration was 2 years ago.
My idea is in theory simple. It is based on the thermal converters used in RF powermeters. Two resistors, two high resolution temperature meters. AC on the first en DC on the second. If both are the same temperature the AC voltage is the same as the DC voltage. But I'm sure some people here have done this in the past. I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done with lightbubs but I never tryed that) I do not mind if it is slow etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
Verzonden met Windows Mail
Van: Bill Gold
Verzonden: ‎maandag‎ ‎25‎ ‎augustus‎ ‎2014 ‎15‎:‎40
Aan: volt-nuts
Randy:
Sorry my fault. You have to use the "RMATH" command to get the various
values stored in the registers. See the "RMATH" command in the User's Guide
for a list of what registers you can read.
I sure haven't found any other guides other than the 4 manuals. User's
Guide, Quick Reference Guide, Calibration Manual, and Assembly Level Repair.
It is just a matter of reading the guides and trying to remember what
commands are available. It took me a lot of time to figure out what
commands I use now. I am sure I am missing other commands that might be
useful.
Bill
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Sunday, August 24, 2014 6:04 PM
Subject: Re: [volt-nuts] 732A and Prologix received
> Bill,
>
> I am trying to figure out the MATH function without much success. I input
> the sequence you said (I looked up the instructions to understand what you
> did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
and
> it shows up on the display when I input BLUE F1. I hit ENTER and it takes
> the 40 measurements and the MATH symbol shows on the display during the
> measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> get a MATH ERR symbol on the display. I tried it a couple of times and
the
> same result so I am doing something wrong. Is there a better source for
> explaining how to do front panel masurements than the User Guide, which
> seems oriented at programming automatic rather than manual measurements.
>
> Randy
>
>
>
> On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
>
> > Randy:
> >
> > The MATH function is accessible from the keypad. I don't have an
IEEE
> > interface right now that works. You can also program the numeric keypad
> > keys to have preprogrammed functions. DEFKEY
> >
> > I have made my own "low thermal" measurement leads from Pomona #4892
> > banana plugs and Belden #9272 wire. Why 9272, because it was handy at
the
> > time. It is tin plated copper, shielded twisted pair 20 ga. I have
plans
> > to do custom cables with 16 ga. bare copper wire that I will twist and
then
> > put a braided shield over it. I simply cannot find what I want so I
will
> > build my own cable. I have done something like this before and it
worked
> > fine. When I get a "round toit".
> >
> > I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> > used
> > in the past to verify my homemade "low thermal" leads as described
above.
> > Frankly I cannot see any difference between using the 1756 cables and my
> > homemade cables once I give them a few minutes for the thermals to go
away.
> > As far as I can tell and measure the differences, if any, are below 0.1
ppm
> > at 10 volts.
> >
> > Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
all
> > adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
as
> > the instability of the readings it is hard to determine which is causing
> > the
> > problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> > following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
> > set
> > the MATH to "Statistics" (store high reading/low reading/ and mean of
the
> > readings) in the registers, the number of readings to "40", the trigger
to
> > "hold" (which keeps the meter from triggering until I press "ENTER" and
> > then
> > trigger the sequence of 40 readings when I push the "ENTER" button. You
> > can
> > do all of this manually from the keypads but since I use this sequence a
> > lot
> > I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
> > Once those 40 readings are finished then you can access the various MATH
> > statistic registers, using the menu, by entering "MATH" and then a 2 for
> > low, a 4 for mean, and 13 for high. Of course you could do all of this
> > through the IEEE also. The 3458A has a very rich set of measurement
> > commands. I am still learning all of them. It depends upon what I am
> > trying to accomplish.
> >
> > Since the 1.018 and 1.0 volt outputs are passive and derived from
> > resistive dividers from the 10 volt, I don't see how they could
contribute
> > to the varying readings you are measuring. I think I would put a short
on
> > the input of the 3458A and manually set the range to 1 volt and then
> > observe
> > the variations that way without the 732A involved. When I do this I see
a
> > variation from low reading to high reading of 0.125 uVolts and then
another
> > 40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
> > of the measurment terminals, GUARD connected doesn't seem to affect the
> > readings. So that is the base noise of the 3458A without the 732A,
> > somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
> > a variation of 0.159 uVolts using the same 40 reading method above. I
> > would
> > use this to determine where your problem might exist. Just having the
> > meter
> > input shorted will point you in the right direction. Meter, cables or
> > 732A.
> >
> > Sorry for the long dissertation. Friends get mad at me for being so
> > detailed sometimes.
> >
> > Bill
> >
> > ----- Original Message -----
> > From: "Randy Evans" <randyevans2688@gmail.com>
> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Sent: Sunday, August 24, 2014 7:22 AM
> > Subject: Re: [volt-nuts] 732A and Prologix received
> >
> >
> > > Bill,
> > >
> > > I have convinced myself that the problem I an seeing is due to
thermals.
> > > If I move the cables (with gold-plated banana plugs) using a small
towel
> > > rather than letting my hand touch the plugs, it is much more stable.
If
> > I
> > > then hold the banana plug with my hand after the reading has
stabilized,
> > > the reading drifts rapidly upward. I am trying to check the stability
of
> > > the reading but I haven't figured out the MATH function yet. I assume
> > this
> > > is a programmed function using GPIB only?
> > >
> > > The stability I am seeing by manually recording readings (using NLPC
of
> > 100
> > > and 1000) is much greater than what you are measuring on your system.
> > Not
> > > sure how to ascertain if it's the 3458A or the 732. The value of the
> > > readings are very different between the two - the 3458 reads about 50
uV
> > > high on the 10 V output and about 12 uV low on the 1V output. Rather
> > large
> > > differences (this is after an ACAL). I need to find some better
cables
> > to
> > > make sure the errors are not due to thermals again.
> > >
> > > Randy
> > >
> > >
> > > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
wrote:
> > >
> > > > Randy:
> > > >
> > > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
fit
> > > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
and
> > > > others
> > > > that are in this size and package. Order from one of the usual
> > electronics
> > > > distributors like Allied, Mouser, Digikey. This is a very common
> > battery
> > > > as
> > > > it is used in a lot of "EXIT" signs so they are lighted when the
power
> > goes
> > > > out. I don't see how the 12V 7AH will fit as they are too large. I
> > guess
> > > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
volt 4
> > AH
> > > > but the terminals are in the wrong place so you will have to
"nibble"
> > out
> > > > the aluminum plate that holds them in the 732A battery pack. You
have
> > to
> > > > be
> > > > careful if you use the 12v 5AH as you will have 4 extra battery
> > connection
> > > > leads to deal with and connect correctly. I would stick with the 6V
> > 4AH.
> > > > New batteries will last around 12 to 14 hours before the "CAL" light
> > goes
> > > > out when AC power is not applied. So shipping to Cal Lab can be a
> > problem
> > > > if it is a distance away, or you have to use a shipper like UPS or
> > FEDEX
> > > > and
> > > > you ship the night before and then use their "Morning delivery" and
the
> > Cal
> > > > Lab is expecting your 732A. Same on the way back to you. Of course
> > you
> > > > could always strap another battery on the 732A and hook it up to the
> > "ext
> > > > power" plug to last longer. I have seen it done. The issue is to
get
> > the
> > > > Cal Lab to charge the extra battery before they ship the 732A back
to
> > you.
> > > >
> > > > When you remove and work on the battery pack always have the AC
> > power
> > > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> > supply
> > > > is working.
> > > >
> > > > The "CAL" light is to indicate that power has not been lost to
the
> > > > Reference Amp or other associated circuits. When the raw supply
> > (battery)
> > > > voltage drops below about 21 volts the "CAL" light will go out.
Below
> > that
> > > > voltage the heater circuits will not work correctly and the 18.6
volt
> > > > regulated supply will not regulate. The requirement is that the
> > Reference
> > > > Amp be kept "alive" at all times to maintain the output voltage that
> > was
> > > > measured at the time of the most recent Calibration or
Certification.
> > When
> > > > the semiconductor junctions are unbiased and cool off when power is
> > lost,
> > > > and then power is restored the result will be a different 10 volts
than
> > > > before the power failure. My experience is that after all of the
years
> > > > that
> > > > these units have been powered up, this won't happen and when power
is
> > lost
> > > > and then restored, even months later, the 732A will come back to
almost
> > > > exactly the same 10 volts as when they lost power, usually with in
0.2
> > PPM
> > > > after 24 hours of "warm up".
> > > >
> > > > What type of hookup leads are you using when measuring the 1
volt
> > > > output? What is the PLC set to? I always use 100 PLC to measure
this.
> > If
> > > > you don't have "low thermal" connection leads you can experience uV
> > changes
> > > > for a minute or more after plugging in the leads due to the
"thermals"
> > > > generated because of the difference in temperature between the
banana
> > jacks
> > > > on the 732A and the banana plugs of the leads. I have found that
even
> > just
> > > > plugging in the lead will generate a thermal difference because of
> > > > difference of temps and some heating due to the physical act of just
> > > > inserting the plug because of friction between the jack and plug (my
> > theory
> > > > at any rate). You have to allow at least a minute or more before
being
> > > > able
> > > > to make a measurement after plugging in the leads. I just measured
the
> > > > variation of the 1 volt output of my 732A and using my 3458A and I
got
> > a
> > > > total difference of 0.159 uV over 40 measurements using 100 PLC on
the
> > 1
> > > > volt range of the 3458A. Using the MATH function and all of the
data
> > you
> > > > can collect. That was after waiting for several minutes after
plugging
> > in
> > > > the leads.
> > > >
> > > > I hope all of this helps.
> > > >
> > > > Bill
> > > >
> > > >
> > > >
> > > > ----- Original Message -----
> > > > From: "Randy Evans" <randyevans2688@gmail.com>
> > > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > > > Sent: Friday, August 22, 2014 9:03 PM
> > > > Subject: Re: [volt-nuts] 732A and Prologix received
> > > >
> > > >
> > > > > Todd,
> > > > >
> > > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> > that
> > I
> > > > > keep topped off and they have very low current draw (~2 to 3 mA at
> > 13.5
> > > > > VDC) when charged and at their float voltage, so I am pretty sure
> > they
> > > > are
> > > > > in good condition. I will look at getting those in the units
after I
> > > > > ascertain the condition of the 732.
> > > > >
> > > > > So now I have a what appears to be a functioning 3458A and a 732A
but
> > > > they
> > > > > slightly disagree. I am like the man with two watches that
disagree
> > on
> > > > the
> > > > > time - which is correct? For the moment, i am only concerned
with
> > > > > stability. The need for absolute accuracy will come later.
> > > > >
> > > > > Randy
> > > > >
> > > > >
> > > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
<tmicallef@gmail.com>
> > > > wrote:
> > > > >
> > > > > > Randy,
> > > > > >
> > > > > > You have two possible choices. It can be configured with 4 x 6v
4Ah
> > > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> > has
> > > > > > modified the battery pack already. A couple of mine needed a
> > nibbler
> > > > tool
> > > > > > to remove enough of the aluminum cover that fits over the tops
of
> > the
> > > > > > batteries. The original cover will short out to the battery tabs
> > > > regardless
> > > > > > of the battery configuration if this is not done.
> > > > > >
> > > > > > You can find larger capacity batteries that will give you
slightly
> > more
> > > > > > battery life. I lost a couple sets of mail-order batteries after
a
> > few
> > > > > > extended outages. I would recommend going with locally bought
> > batteries
> > > > > > instead of the cheaper mail order. My local Batteries Plus will
> > > > typically
> > > > > > have some warranty if I remember correctly. Moving forward I
will
> > only
> > > > use
> > > > > > 2 12v batteries and pre-charge them on a battery charger to
> > equalize
> > > > them
> > > > > > before putting them in the 732A. I think the cheap batteries did
> > not
> > > > > > discharge equally, and would not recover when power was applied.
> > > > > >
> > > > > > Inspect the back plane for damaged traces and look at the
> > capacitors. I
> > > > had
> > > > > > a few that looked questionable. So far, I have replaced all the
big
> > > > caps
> > > > on
> > > > > > the pre-regulator and regulator boards. My feeling is that once
> > these
> > > > go
> > > > > > online, they should run as long as possible between repairs.
> > > > > >
> > > > > > The battery charger circuit may need adjusting. I tweaked mine
and
> > it
> > > > > > seemed to work fine.
> > > > > >
> > > > > > Todd
> > > > > >
> > > > > >
> > > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > > randyevans2688@gmail.com>
> > > > > > wrote:
> > > > > >
> > > > > > > I received my Fluke 732A today. Just powered it up but it
needs
> > new
> > > > > > > batteries. Any suggestions for sources (I haven't opened up
the
> > unit
> > > > > > yet -
> > > > > > > I want to make sure it works before doing that). Also
received
> > the
> > > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> > data
> > > > > > dumper
> > > > > > > program to get the CAL data from my 3458A. Should be a busy
> > weekend.
> > > > > > >
> > > > > > > Randy
> > > > > > > _______________________________________________
> > > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > > To unsubscribe, go to
> > > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > > and follow the instructions there.
> > > > > > >
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > > To unsubscribe, go to
> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
D
Don@True-Cal
Mon, Aug 25, 2014 3:02 PM
Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal.
The NI website had this chart that quantifies the Seebeck voltage best:
"When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients."
Junction µV/°C
Copper-Copper <0.3
Copper-Gold 0.5
Copper-Silver 0.5
Copper-Brass 3
Copper-Nickel 10
Copper-Lead-Tin Solder 1-3
Copper-Aluminum 5
Copper-Kovar 40
Copper-Copper Oxide >500
Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Monday, August 25, 2014 5:37 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them.
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal.
The NI website had this chart that quantifies the Seebeck voltage best:
"When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients."
Junction µV/°C
Copper-Copper <0.3
Copper-Gold 0.5
Copper-Silver 0.5
Copper-Brass 3
Copper-Nickel 10
Copper-Lead-Tin Solder 1-3
Copper-Aluminum 5
Copper-Kovar 40
Copper-Copper Oxide >500
Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Monday, August 25, 2014 5:37 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them.
> Gesendet: Montag, 25. August 2014 um 06:33 Uhr
> Von: "Orin Eman" <orin.eman@gmail.com>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com>
> wrote:
>
> > Randy & all,
> >
> > You have correctly concluded that some (maybe not all) of your measurement
> > problem is thermal EMF being added or subtracted in series within your
> > measurement interconnect. This thermal EMF is generated at the junction of
> > dissimilar metals when accompanied with thermal gradients between the test
> > lead and device terminals. You have to eliminate both the dissimilarity of
> > the metal junctions as well as minimize the thermal differences. The
> > terminals of the 3458A as well as the 732A are Beryllium Copper so you want
> > to use the same test lead terminals. Forget the typical Tin plated lugs or
> > even Gold plated as both are not Beryllium Copper and constitute dissimilar
> > metals. The best solution (as usually the most expensive) is to use a set
> > of
> > Fluke 5440A-7005 (48") cables. I also have just as good results using the
> > much more flexible Pomona 11174A (lugs end always stay connected to the
> > 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
> > has the added Guard built in but be sure to also use a Guard lead with the
> > Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
> > cables is usually not a good idea because the lead wire to terminal also
> > constitutes just as critical of junction. The above cables use Tellurium
> > Copper wire which is usually hard to find and hard to crimp properly and
> > NEVER solder.
> >
>
>
> 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade
> to banana cables are available (5295-36 at Mouser et al) and claim that
> they are designed to minimize thermal EMFs. Datasheet is here:
> http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these
> as an alternative?
>
> Orin.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
A
acbern@gmx.de
Mon, Aug 25, 2014 3:48 PM
well, your last point is the issue, how can you have a temperature difference within a few microns of material in said connections. theory is one thing, but in reality it does not happen due to the givens of the setup.
therefore in practice it is irrelevant if the wire is silver or gold plated or pure copper. otherwise the gold plated spades and tellurium copper posts from pomona and others would be nonsense. and other than the mysterious fluke wire I have never seen a tellurium-copper wire from any wire manufacturer.
Gesendet: Montag, 25. August 2014 um 17:02 Uhr
Von: "Don@True-Cal" truecalservices@gmail.com
An: "'Discussion of precise voltage measurement'" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal.
The NI website had this chart that quantifies the Seebeck voltage best:
"When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients."
Junction µV/°C
Copper-Copper <0.3
Copper-Gold 0.5
Copper-Silver 0.5
Copper-Brass 3
Copper-Nickel 10
Copper-Lead-Tin Solder 1-3
Copper-Aluminum 5
Copper-Kovar 40
Copper-Copper Oxide >500
Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Monday, August 25, 2014 5:37 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them.
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
well, your last point is the issue, how can you have a temperature difference within a few microns of material in said connections. theory is one thing, but in reality it does not happen due to the givens of the setup.
therefore in practice it is irrelevant if the wire is silver or gold plated or pure copper. otherwise the gold plated spades and tellurium copper posts from pomona and others would be nonsense. and other than the mysterious fluke wire I have never seen a tellurium-copper wire from any wire manufacturer.
> Gesendet: Montag, 25. August 2014 um 17:02 Uhr
> Von: "Don@True-Cal" <truecalservices@gmail.com>
> An: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal.
>
> The NI website had this chart that quantifies the Seebeck voltage best:
> "When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients."
>
> Junction µV/°C
> Copper-Copper <0.3
> Copper-Gold 0.5
> Copper-Silver 0.5
> Copper-Brass 3
> Copper-Nickel 10
> Copper-Lead-Tin Solder 1-3
> Copper-Aluminum 5
> Copper-Kovar 40
> Copper-Copper Oxide >500
>
> Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences.
>
> Don Johnson
>
> -----Original Message-----
> From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
> Sent: Monday, August 25, 2014 5:37 AM
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] 732A and Prologix received
>
> I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
> so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
> that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
> 1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them.
>
>
>
> > Gesendet: Montag, 25. August 2014 um 06:33 Uhr
> > Von: "Orin Eman" <orin.eman@gmail.com>
> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Betreff: Re: [volt-nuts] 732A and Prologix received
> >
> > On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com>
> > wrote:
> >
> > > Randy & all,
> > >
> > > You have correctly concluded that some (maybe not all) of your measurement
> > > problem is thermal EMF being added or subtracted in series within your
> > > measurement interconnect. This thermal EMF is generated at the junction of
> > > dissimilar metals when accompanied with thermal gradients between the test
> > > lead and device terminals. You have to eliminate both the dissimilarity of
> > > the metal junctions as well as minimize the thermal differences. The
> > > terminals of the 3458A as well as the 732A are Beryllium Copper so you want
> > > to use the same test lead terminals. Forget the typical Tin plated lugs or
> > > even Gold plated as both are not Beryllium Copper and constitute dissimilar
> > > metals. The best solution (as usually the most expensive) is to use a set
> > > of
> > > Fluke 5440A-7005 (48") cables. I also have just as good results using the
> > > much more flexible Pomona 11174A (lugs end always stay connected to the
> > > 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
> > > has the added Guard built in but be sure to also use a Guard lead with the
> > > Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
> > > cables is usually not a good idea because the lead wire to terminal also
> > > constitutes just as critical of junction. The above cables use Tellurium
> > > Copper wire which is usually hard to find and hard to crimp properly and
> > > NEVER solder.
> > >
> >
> >
> > 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade
> > to banana cables are available (5295-36 at Mouser et al) and claim that
> > they are designed to minimize thermal EMFs. Datasheet is here:
> > http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these
> > as an alternative?
> >
> > Orin.
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
MS
Mike S
Mon, Aug 25, 2014 4:02 PM
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
Silver or Gold plating on the terminal or wire will introduce the
undesirable dissimilar metal properties, both at the plating
junction and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction,
since both junctions are (under normal conditions) at essentially the
same temperature (e.g. there's a copper-gold thermocouple, the minimal
thermal resistance of a micron of gold on the contact(s), then a
gold-copper thermocouple). It seems to me that the improved consistency
of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the
concept is the same. Unless there is heat flowing through the entire
assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they
simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out
the same - a copper-nickel-gold-nickel-copper connection is completely
offset. It's when the offsetting thermocouples occur across a
temperature gradient that you have problems.
--
Mike
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
> Silver or Gold plating on the terminal or wire will introduce the
> undesirable dissimilar metal properties, both at the plating
> junction and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction,
since both junctions are (under normal conditions) at essentially the
same temperature (e.g. there's a copper-gold thermocouple, the minimal
thermal resistance of a micron of gold on the contact(s), then a
gold-copper thermocouple). It seems to me that the improved consistency
of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the
concept is the same. Unless there is heat flowing through the entire
assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they
simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out
the same - a copper-nickel-gold-nickel-copper connection is completely
offset. It's when the offsetting thermocouples occur across a
temperature gradient that you have problems.
--
Mike
CB
Charles Black
Mon, Aug 25, 2014 4:50 PM
Hi Adrian,
Although I have not used the Pomona spade lugs or a nano-voltmeter, my
experience is consistent with yours otherwise.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire to short the top four input terminals. This
is very convenient since it is so easy to duplicate in the field and
makes simple inexpensive test leads best for high precision
measurements. In order for the 3458A to make full accuracy measurements
(8 digit) NPLC must be set to 1000 (according to the User's Guide). If
you use a lower NPLC value there is a table in the User's Guide that can
be used to determine how accurate your measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv +/- 10
pV Equilibration time 5 minutes. 14 gauge per Calibration
Manual.
"U" heavy wire -0.00021mv +/-
10pV Equil. time 5 minutes. Used the through holes in the
Input banana posts only.
Copper wire -0.00019mv +/-
10pV NAPA PVC covered automobile wire at same contact
points as CAL shunt
Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5
minutes. Gold plated ground plate from my Datron 4910
Copper wire -0.00019mv +/-
10Pv Equil. time 2 seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv +/- 10pV
Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv +/-
30pV Equil. time 20 minutes. My best "Perfect" gold plated
plugs with copper wire.
Charlie
On 8/25/2014 3:36 AM, acbern@gmx.de wrote:
I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.
Randy & all,
You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.
Hi Adrian,
Although I have not used the Pomona spade lugs or a nano-voltmeter, my
experience is consistent with yours otherwise.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire to short the top four input terminals. This
is very convenient since it is so easy to duplicate in the field and
makes simple inexpensive test leads best for high precision
measurements. In order for the 3458A to make full accuracy measurements
(8 digit) NPLC must be set to 1000 (according to the User's Guide). If
you use a lower NPLC value there is a table in the User's Guide that can
be used to determine how accurate your measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv +/- 10
pV Equilibration time 5 minutes. 14 gauge per Calibration
Manual.
"U" heavy wire -0.00021mv +/-
10pV Equil. time 5 minutes. Used the through holes in the
Input banana posts only.
Copper wire -0.00019mv +/-
10pV NAPA PVC covered automobile wire at same contact
points as CAL shunt
Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5
minutes. Gold plated ground plate from my Datron 4910
Copper wire -0.00019mv +/-
10Pv Equil. time 2 seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv +/- 10pV
Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv +/-
30pV Equil. time 20 minutes. My best "Perfect" gold plated
plugs with copper wire.
Charlie
On 8/25/2014 3:36 AM, acbern@gmx.de wrote:
> I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
> so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
> that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
> 1.copper-tellurium post from e.g. 8 digit calibrator
> 2.crimped copper spade, gold plated
> 3.silver plated tsp copper cable
> 4a.crimped copper spade to copper-tellurium post or
> 4b.soldered copper connector(34420)
> my consistent results over more than a year using them.
>
>
>
>> Gesendet: Montag, 25. August 2014 um 06:33 Uhr
>> Von: "Orin Eman" <orin.eman@gmail.com>
>> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Betreff: Re: [volt-nuts] 732A and Prologix received
>>
>> On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com>
>> wrote:
>>
>>> Randy & all,
>>>
>>> You have correctly concluded that some (maybe not all) of your measurement
>>> problem is thermal EMF being added or subtracted in series within your
>>> measurement interconnect. This thermal EMF is generated at the junction of
>>> dissimilar metals when accompanied with thermal gradients between the test
>>> lead and device terminals. You have to eliminate both the dissimilarity of
>>> the metal junctions as well as minimize the thermal differences. The
>>> terminals of the 3458A as well as the 732A are Beryllium Copper so you want
>>> to use the same test lead terminals. Forget the typical Tin plated lugs or
>>> even Gold plated as both are not Beryllium Copper and constitute dissimilar
>>> metals. The best solution (as usually the most expensive) is to use a set
>>> of
>>> Fluke 5440A-7005 (48") cables. I also have just as good results using the
>>> much more flexible Pomona 11174A (lugs end always stay connected to the
>>> 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
>>> has the added Guard built in but be sure to also use a Guard lead with the
>>> Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
>>> cables is usually not a good idea because the lead wire to terminal also
>>> constitutes just as critical of junction. The above cables use Tellurium
>>> Copper wire which is usually hard to find and hard to crimp properly and
>>> NEVER solder.
>>>
>>
>> 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade
>> to banana cables are available (5295-36 at Mouser et al) and claim that
>> they are designed to minimize thermal EMFs. Datasheet is here:
>> http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these
>> as an alternative?
>>
>> Orin.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
>
DM
Dave M
Mon, Aug 25, 2014 6:38 PM
Well, you sort of answered your own question. The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done.
Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer. Leaving the reference voltage setting alone, a DC voltage is
input into the unit, and the DC source is adjusted for a null on the
galvanometer. At that point, the AC voltage source is equal to that of the
DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you have to
have a converter for each range of voltages that you need to measure. The
thermal converters used with this type of transfer standard isn't great (50
MHz or so typical), but their accuracy far surpasses that of the thermistor
type sensors.
There are other brands and models of thermal transfer standards, but I have
a Fluke model 540 and a few thermal converters. That's why I referred you
to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two AC+DC
TRMS 7,5 digit meters but the last calibration was 2 years ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for 1V,
10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
Well, you sort of answered your own question. The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done.
Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer. Leaving the reference voltage setting alone, a DC voltage is
input into the unit, and the DC source is adjusted for a null on the
galvanometer. At that point, the AC voltage source is equal to that of the
DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you have to
have a converter for each range of voltages that you need to measure. The
thermal converters used with this type of transfer standard isn't great (50
MHz or so typical), but their accuracy far surpasses that of the thermistor
type sensors.
There are other brands and models of thermal transfer standards, but I have
a Fluke model 540 and a few thermal converters. That's why I referred you
to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
> Is there a way to link an AC voltage to a DC source for compare. I
> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
> against standardcells. But for AC I can not do that. I have two AC+DC
> TRMS 7,5 digit meters but the last calibration was 2 years ago.
>
> My idea is in theory simple. It is based on the thermal converters
> used in RF powermeters. Two resistors, two high resolution
> temperature meters. AC on the first en DC on the second. If both are
> the same temperature the AC voltage is the same as the DC voltage.
> But I'm sure some people here have done this in the past. I would
> like to use it for 50 to 100 kHz (or less) and something like for 1V,
> 10V and 100V (and use several resistors/heaters.)
>
> Or mabey there is an other way to convert AC (for RF it can be done
> with lightbubs but I never tryed that) I do not mind if it is slow
> etc, I like this sort of experiments. You can learn a lot from it.
>
> Fred, pa4tim
A
acbern@gmx.de
Mon, Aug 25, 2014 7:05 PM
fred,
generally you raise a good point, I had the same issue of calibrating an ac voltage to a high level of accuracy. you need this e.g. to validate the self.cal of a 3458a or other precison stuff like the 8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need and a second one on range above. build a set of resistor range extenders (rf type with appropriate connectors and housings) to expand the range to where you need to be max. get one of the thermal converter calibrated (the higher one usually, and you need to havr good cal lab, should be <10ppm accuracy) and use it to calibrate the rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for thermal converters! at higher frequencies, due to reflections and stray capacitance/inductance influences, the accuracy decreases. the resistor range extenders though, if build up correctly, only have a few ppm impact (there is a paper from nist on that, but this is only typical). you can calibrate all converters to the one you got externally calibrated. do some research in the web, when you do the calibration, you need to determine the so-called constant N. then do an ac, dc+, ac, dc-, ac measurement between the the two and establish the deviation, also establish the error propagation. the end result will be a set of highly precise (low inaccuracies9 thermal converters good enough to calibrate a 3458a an better devices. if you want to spend the money, you could also buy a set of converters/range resistors (with/without a 540), that typically is a few k altogether, while a single device sometimes is available for below 100 bucks. you need to have a stable 7.5 digit nanovoltmeter though for the measurements of the tvcs (34420a or 2182 typically ) and precision (stable) dc and ac sources. but in the end, all you need is a single calibrated thermal converter.
adrian
Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done.
Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer. Leaving the reference voltage setting alone, a DC voltage is
input into the unit, and the DC source is adjusted for a null on the
galvanometer. At that point, the AC voltage source is equal to that of the
DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you have to
have a converter for each range of voltages that you need to measure. The
thermal converters used with this type of transfer standard isn't great (50
MHz or so typical), but their accuracy far surpasses that of the thermistor
type sensors.
There are other brands and models of thermal transfer standards, but I have
a Fluke model 540 and a few thermal converters. That's why I referred you
to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two AC+DC
TRMS 7,5 digit meters but the last calibration was 2 years ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for 1V,
10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
fred,
generally you raise a good point, I had the same issue of calibrating an ac voltage to a high level of accuracy. you need this e.g. to validate the self.cal of a 3458a or other precison stuff like the 8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need and a second one on range above. build a set of resistor range extenders (rf type with appropriate connectors and housings) to expand the range to where you need to be max. get one of the thermal converter calibrated (the higher one usually, and you need to havr good cal lab, should be <10ppm accuracy) and use it to calibrate the rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for thermal converters! at higher frequencies, due to reflections and stray capacitance/inductance influences, the accuracy decreases. the resistor range extenders though, if build up correctly, only have a few ppm impact (there is a paper from nist on that, but this is only typical). you can calibrate all converters to the one you got externally calibrated. do some research in the web, when you do the calibration, you need to determine the so-called constant N. then do an ac, dc+, ac, dc-, ac measurement between the the two and establish the deviation, also establish the error propagation. the end result will be a set of highly precise (low inaccuracies9 thermal converters good enough to calibrate a 3458a an better devices. if you want to spend the money, you could also buy a set of converters/range resistors (with/without a 540), that typically is a few k altogether, while a single device sometimes is available for below 100 bucks. you need to have a stable 7.5 digit nanovoltmeter though for the measurements of the tvcs (34420a or 2182 typically ) and precision (stable) dc and ac sources. but in the end, all you need is a single calibrated thermal converter.
adrian
> Gesendet: Montag, 25. August 2014 um 18:38 Uhr
> Von: "Dave M" <dgminala@mediacombb.net>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] AC calibration
>
> Well, you sort of answered your own question. The equipment is called a
> Thermal Transfer Standard, but instead of thermistors, it uses a
> thermocouple. Look at the manual for the Fluke 540B
> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done.
> Basically, the AC source is input into the transfer standard, and the
> standard's internal reference voltage is adjusted for a null on the
> galvanometer. Leaving the reference voltage setting alone, a DC voltage is
> input into the unit, and the DC source is adjusted for a null on the
> galvanometer. At that point, the AC voltage source is equal to that of the
> DC voltage source.
>
> Ther are thermocouple-type thermal converters used for RF voltage
> measurements with the transfer standard. They aren't cheap, and you have to
> have a converter for each range of voltages that you need to measure. The
> thermal converters used with this type of transfer standard isn't great (50
> MHz or so typical), but their accuracy far surpasses that of the thermistor
> type sensors.
>
> There are other brands and models of thermal transfer standards, but I have
> a Fluke model 540 and a few thermal converters. That's why I referred you
> to the manual for it.
>
> Cheers,
> Dave M
>
>
> pa4tim@gmail.com wrote:
> > Is there a way to link an AC voltage to a DC source for compare. I
> > can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
> > against standardcells. But for AC I can not do that. I have two AC+DC
> > TRMS 7,5 digit meters but the last calibration was 2 years ago.
> >
> > My idea is in theory simple. It is based on the thermal converters
> > used in RF powermeters. Two resistors, two high resolution
> > temperature meters. AC on the first en DC on the second. If both are
> > the same temperature the AC voltage is the same as the DC voltage.
> > But I'm sure some people here have done this in the past. I would
> > like to use it for 50 to 100 kHz (or less) and something like for 1V,
> > 10V and 100V (and use several resistors/heaters.)
> >
> > Or mabey there is an other way to convert AC (for RF it can be done
> > with lightbubs but I never tryed that) I do not mind if it is slow
> > etc, I like this sort of experiments. You can learn a lot from it.
> >
> > Fred, pa4tim
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
CH
Chuck Harris
Mon, Aug 25, 2014 7:12 PM
Look up the phrase "AC thermal transfer standard".
Using a heater/thermocouple element in a vacuum is the tried
and true way of linking an AC voltage to a DC reference.
-Chuck Harris
OBTW, trimming your quoted posts is considered friendly.
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I can check my
calibrators (like a Fluke 332, 760 , 731 and a Philips) against standardcells.
But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last
calibration was 2 years ago.
My idea is in theory simple. It is based on the thermal converters used in RF
powermeters. Two resistors, two high resolution temperature meters. AC on the
first en DC on the second. If both are the same temperature the AC voltage is the
same as the DC voltage. But I'm sure some people here have done this in the past.
I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V
and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done with lightbubs
but I never tryed that) I do not mind if it is slow etc, I like this sort of
experiments. You can learn a lot from it.
Fred, pa4tim
Look up the phrase "AC thermal transfer standard".
Using a heater/thermocouple element in a vacuum is the tried
and true way of linking an AC voltage to a DC reference.
-Chuck Harris
OBTW, trimming your quoted posts is considered friendly.
pa4tim@gmail.com wrote:
> Is there a way to link an AC voltage to a DC source for compare. I can check my
> calibrators (like a Fluke 332, 760 , 731 and a Philips) against standardcells.
> But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last
> calibration was 2 years ago.
>
>
> My idea is in theory simple. It is based on the thermal converters used in RF
> powermeters. Two resistors, two high resolution temperature meters. AC on the
> first en DC on the second. If both are the same temperature the AC voltage is the
> same as the DC voltage. But I'm sure some people here have done this in the past.
> I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V
> and 100V (and use several resistors/heaters.)
>
>
> Or mabey there is an other way to convert AC (for RF it can be done with lightbubs
> but I never tryed that) I do not mind if it is slow etc, I like this sort of
> experiments. You can learn a lot from it.
>
>
>
>
> Fred, pa4tim
CB
Charles Black
Mon, Aug 25, 2014 7:28 PM
I just realized that all my measurement tolerances are nV not pV. Sorry
about not checking units first.
Charlie
On 8/25/2014 9:50 AM, Charles Black wrote:
Hi Adrian,
Although I have not used the Pomona spade lugs or a nano-voltmeter, my
experience is consistent with yours otherwise.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough
information to deduce that lowest input short voltage is going to be a
copper wire since the meter is set to zero volts during calibration
using a heavy gauge (14 to 16) copper wire to short the top four input
terminals. This is very convenient since it is so easy to duplicate in
the field and makes simple inexpensive test leads best for high
precision measurements. In order for the 3458A to make full accuracy
measurements (8 digit) NPLC must be set to 1000 (according to the
User's Guide). If you use a lower NPLC value there is a table in the
User's Guide that can be used to determine how accurate your
measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has
been about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv +/- 10
pV Equilibration time 5 minutes. 14 gauge per Calibration
Manual.
"U" heavy wire -0.00021mv +/-
10pV Equil. time 5 minutes. Used the through holes in the
Input banana posts only.
Copper wire -0.00019mv +/-
10pV NAPA PVC covered automobile wire at same contact
points as CAL shunt
Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5
minutes. Gold plated ground plate from my Datron 4910
Copper wire -0.00019mv +/-
10Pv Equil. time 2 seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv +/- 10pV
Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv +/-
30pV Equil. time 20 minutes. My best "Perfect" gold
plated plugs with copper wire.
Charlie
On 8/25/2014 3:36 AM, acbern@gmx.de wrote:
I have used the pomona spades, mainly to interface the low emf pomona
banana cables to binding posts. I have stopped this, reasons being,
they are large and worse, that the pomona spring loaded insulation
tube that covers the banana plug conductor uses such a strong spring
that slowly the plug works its way out of the spade. this btw also
happend to me when I used the pomona low emf binding posts together
with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded
pair of high grade teflon/kapton silver plated copper cable with gold
plated copper spades (crimped). I use them not only with the 3458a
but also with nanovolt meters. these have higher resolution and
accuracy in low level measurements than the 3458a. emf voltages were
never an issue with these cables if properly used. I have posted some
results doing 34420a stabilty measurements on the pmel forum, and the
results are convincing (purpose was actually not to test the cables
but the stability of the 34420a, but the emf issue is a part of this
of course. we use the 34420a to do low voltage precision measurements
on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr
with his comments about copper telurium as cable and spade material
and so on. this material, as stated here many times, is used because
it is machinable, for copper spades one would not use it. the 34420a
factory cable uses copper cable and copper spades, not
telurium-copper. if there was a problem, it would be worse with the
34420a than with the 3458a because of its low level ranges. and
again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.
Randy & all,
You have correctly concluded that some (maybe not all) of your
measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the
junction of
dissimilar metals when accompanied with thermal gradients between
the test
lead and device terminals. You have to eliminate both the
dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so
you want
to use the same test lead terminals. Forget the typical Tin plated
lugs or
even Gold plated as both are not Beryllium Copper and constitute
dissimilar
metals. The best solution (as usually the most expensive) is to use
a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results
using the
much more flexible Pomona 11174A (lugs end always stay connected to
the
732A) or 11058A with more convenient shielded banana plugs. The
Fluke cable
has the added Guard built in but be sure to also use a Guard lead
with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal
also
constitutes just as critical of junction. The above cables use
Tellurium
Copper wire which is usually hard to find and hard to crimp
properly and
NEVER solder.
I just realized that all my measurement tolerances are nV not pV. Sorry
about not checking units first.
Charlie
On 8/25/2014 9:50 AM, Charles Black wrote:
> Hi Adrian,
>
> Although I have not used the Pomona spade lugs or a nano-voltmeter, my
> experience is consistent with yours otherwise.
>
> Anyone who has calibrated (CAL 0 anyway) a 3458A has enough
> information to deduce that lowest input short voltage is going to be a
> copper wire since the meter is set to zero volts during calibration
> using a heavy gauge (14 to 16) copper wire to short the top four input
> terminals. This is very convenient since it is so easy to duplicate in
> the field and makes simple inexpensive test leads best for high
> precision measurements. In order for the 3458A to make full accuracy
> measurements (8 digit) NPLC must be set to 1000 (according to the
> User's Guide). If you use a lower NPLC value there is a table in the
> User's Guide that can be used to determine how accurate your
> measurements are going to be.
>
> Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has
> been about a year since I clid my last CAL 0 so it was going to be
> interesting at least for me. The first shunt was my test "U' shaped
> shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
>
> Calibration shunt -0.00021mv +/- 10
> pV Equilibration time 5 minutes. 14 gauge per Calibration
> Manual.
> "U" heavy wire -0.00021mv +/-
> 10pV Equil. time 5 minutes. Used the through holes in the
> Input banana posts only.
> Copper wire -0.00019mv +/-
> 10pV NAPA PVC covered automobile wire at same contact
> points as CAL shunt
> Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5
> minutes. Gold plated ground plate from my Datron 4910
> Copper wire -0.00019mv +/-
> 10Pv Equil. time 2 seconds. Used the banana through holes.
> STP 2 meter test lead -0.00021mv +/- 10pV
> Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation.
> Banana plugs -0.00021mv +/-
> 30pV Equil. time 20 minutes. My best "Perfect" gold
> plated plugs with copper wire.
>
> Charlie
>
>
> On 8/25/2014 3:36 AM, acbern@gmx.de wrote:
>> I have used the pomona spades, mainly to interface the low emf pomona
>> banana cables to binding posts. I have stopped this, reasons being,
>> they are large and worse, that the pomona spring loaded insulation
>> tube that covers the banana plug conductor uses such a strong spring
>> that slowly the plug works its way out of the spade. this btw also
>> happend to me when I used the pomona low emf binding posts together
>> with the pomona low emf banana cables. overall I m not happy with these.
>> so, due to lack of options, I changed to self-made twisted shielded
>> pair of high grade teflon/kapton silver plated copper cable with gold
>> plated copper spades (crimped). I use them not only with the 3458a
>> but also with nanovolt meters. these have higher resolution and
>> accuracy in low level measurements than the 3458a. emf voltages were
>> never an issue with these cables if properly used. I have posted some
>> results doing 34420a stabilty measurements on the pmel forum, and the
>> results are convincing (purpose was actually not to test the cables
>> but the stability of the 34420a, but the emf issue is a part of this
>> of course. we use the 34420a to do low voltage precision measurements
>> on thermal converters where the full scale signal sometimes is 1mV).
>> that btw also relates to don's statements below, I do not concurr
>> with his comments about copper telurium as cable and spade material
>> and so on. this material, as stated here many times, is used because
>> it is machinable, for copper spades one would not use it. the 34420a
>> factory cable uses copper cable and copper spades, not
>> telurium-copper. if there was a problem, it would be worse with the
>> 34420a than with the 3458a because of its low level ranges. and
>> again, I have not seen any problems in a chain of (output to input):
>> 1.copper-tellurium post from e.g. 8 digit calibrator
>> 2.crimped copper spade, gold plated
>> 3.silver plated tsp copper cable
>> 4a.crimped copper spade to copper-tellurium post or
>> 4b.soldered copper connector(34420)
>> my consistent results over more than a year using them.
>>
>>
>>
>>> Gesendet: Montag, 25. August 2014 um 06:33 Uhr
>>> Von: "Orin Eman" <orin.eman@gmail.com>
>>> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>>> Betreff: Re: [volt-nuts] 732A and Prologix received
>>>
>>> On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal
>>> <truecalservices@gmail.com>
>>> wrote:
>>>
>>>> Randy & all,
>>>>
>>>> You have correctly concluded that some (maybe not all) of your
>>>> measurement
>>>> problem is thermal EMF being added or subtracted in series within your
>>>> measurement interconnect. This thermal EMF is generated at the
>>>> junction of
>>>> dissimilar metals when accompanied with thermal gradients between
>>>> the test
>>>> lead and device terminals. You have to eliminate both the
>>>> dissimilarity of
>>>> the metal junctions as well as minimize the thermal differences. The
>>>> terminals of the 3458A as well as the 732A are Beryllium Copper so
>>>> you want
>>>> to use the same test lead terminals. Forget the typical Tin plated
>>>> lugs or
>>>> even Gold plated as both are not Beryllium Copper and constitute
>>>> dissimilar
>>>> metals. The best solution (as usually the most expensive) is to use
>>>> a set
>>>> of
>>>> Fluke 5440A-7005 (48") cables. I also have just as good results
>>>> using the
>>>> much more flexible Pomona 11174A (lugs end always stay connected to
>>>> the
>>>> 732A) or 11058A with more convenient shielded banana plugs. The
>>>> Fluke cable
>>>> has the added Guard built in but be sure to also use a Guard lead
>>>> with the
>>>> Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
>>>> cables is usually not a good idea because the lead wire to terminal
>>>> also
>>>> constitutes just as critical of junction. The above cables use
>>>> Tellurium
>>>> Copper wire which is usually hard to find and hard to crimp
>>>> properly and
>>>> NEVER solder.
>>>>
>>>
>>> 11058A and 11174A are discontinued at Keysight. However, Pomona
>>> 5295 spade
>>> to banana cables are available (5295-36 at Mouser et al) and claim that
>>> they are designed to minimize thermal EMFs. Datasheet is here:
>>> http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on
>>> these
>>> as an alternative?
>>>
>>> Orin.
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>>
>>
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
>
DM
Dave M
Mon, Aug 25, 2014 7:32 PM
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Dave M
acbern@gmx.de wrote:
fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be <10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.
adrian
Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer. Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer. At that point,
the AC voltage source is equal to that of the DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you
have to have a converter for each range of voltages that you need to
measure. The thermal converters used with this type of transfer
standard isn't great (50 MHz or so typical), but their accuracy far
surpasses that of the thermistor type sensors.
There are other brands and models of thermal transfer standards, but
I have a Fluke model 540 and a few thermal converters. That's why I
referred you to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two
AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for
1V, 10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
When the people fear the government, there is tyranny. When the
government fears the people, there is liberty -- Thomas Jefferson
Dave M
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Dave M
acbern@gmx.de wrote:
> fred,
> generally you raise a good point, I had the same issue of calibrating
> an ac voltage to a high level of accuracy. you need this e.g. to
> validate the self.cal of a 3458a or other precison stuff like the
> 8506a0.
>
> what i would recommend to do if you want to keep costs down is:
> in a nutshell, get a thermal converter in the lowest range you need
> and a second one on range above. build a set of resistor range
> extenders (rf type with appropriate connectors and housings) to
> expand the range to where you need to be max. get one of the thermal
> converter calibrated (the higher one usually, and you need to havr
> good cal lab, should be <10ppm accuracy) and use it to calibrate the
> rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
> thermal converters! at higher frequencies, due to reflections and
> stray capacitance/inductance influences, the accuracy decreases. the
> resistor range extenders though, if build up correctly, only have a
> few ppm impact (there is a paper from nist on that, but this is only
> typical). you can calibrate all converters to the one you got
> externally calibrated. do some research in the web, when you do the
> calibration, you need to determine the so-called constant N. then do
> an ac, dc+, ac, dc-, ac measurement between the the two and establish
> the deviation, also establish the error propagation. the end result
> will be a set of highly precise (low inaccuracies9 thermal converters
> good enough to calibrate a 3458a an better devices. if you want to
> spend the money, you could also buy a set of converters/range
> resistors (with/without a 540), that typically is a few k altogether,
> while a single device sometimes is available for below 100 bucks. you
> need to have a stable 7.5 digit nanovoltmeter though for the
> measurements of the tvcs (34420a or 2182 typically ) and precision
> (stable) dc and ac sources. but in the end, all you need is a single
> calibrated thermal converter.
>
> adrian
>
>
>
>> Gesendet: Montag, 25. August 2014 um 18:38 Uhr
>> Von: "Dave M" <dgminala@mediacombb.net>
>> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Betreff: Re: [volt-nuts] AC calibration
>>
>> Well, you sort of answered your own question. The equipment is
>> called a Thermal Transfer Standard, but instead of thermistors, it
>> uses a thermocouple. Look at the manual for the Fluke 540B
>> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
>> it's done. Basically, the AC source is input into the transfer
>> standard, and the standard's internal reference voltage is adjusted
>> for a null on the galvanometer. Leaving the reference voltage
>> setting alone, a DC voltage is input into the unit, and the DC
>> source is adjusted for a null on the galvanometer. At that point,
>> the AC voltage source is equal to that of the DC voltage source.
>>
>> Ther are thermocouple-type thermal converters used for RF voltage
>> measurements with the transfer standard. They aren't cheap, and you
>> have to have a converter for each range of voltages that you need to
>> measure. The thermal converters used with this type of transfer
>> standard isn't great (50 MHz or so typical), but their accuracy far
>> surpasses that of the thermistor type sensors.
>>
>> There are other brands and models of thermal transfer standards, but
>> I have a Fluke model 540 and a few thermal converters. That's why I
>> referred you to the manual for it.
>>
>> Cheers,
>> Dave M
>>
>>
>> pa4tim@gmail.com wrote:
>>> Is there a way to link an AC voltage to a DC source for compare. I
>>> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
>>> against standardcells. But for AC I can not do that. I have two
>>> AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
>>> ago.
>>>
>>> My idea is in theory simple. It is based on the thermal converters
>>> used in RF powermeters. Two resistors, two high resolution
>>> temperature meters. AC on the first en DC on the second. If both are
>>> the same temperature the AC voltage is the same as the DC voltage.
>>> But I'm sure some people here have done this in the past. I would
>>> like to use it for 50 to 100 kHz (or less) and something like for
>>> 1V, 10V and 100V (and use several resistors/heaters.)
>>>
>>> Or mabey there is an other way to convert AC (for RF it can be done
>>> with lightbubs but I never tryed that) I do not mind if it is slow
>>> etc, I like this sort of experiments. You can learn a lot from it.
>>>
>>> Fred, pa4tim
>>
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
>> the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
When the people fear the government, there is tyranny. When the
government fears the people, there is liberty -- Thomas Jefferson
Dave M
BG
Bill Gold
Mon, Aug 25, 2014 10:09 PM
There is one more thing that enters into this discussion and that is
"reversal errors" on the DC. The complicates the transfer somewhat. AC is
always going "+" and "-". DC is in one direction so you have to then
reverse the voltage to the Thermal Transfer Standard and then take the
average of the two readings. That is why there is a "Reversal" switch on
the 540B. When you are using a fixed voltage High Frequency Thermal
Converter you need an external DC reversal switch in addition to other
equipment. You also need an AC/DC transfer switch so that you don't have to
disconnect the AC source and then hook up the DC source manually. See the
540B again.
All Thermal Transfer Standards have some "reversal" error. This is
controlled by the internal construction of the unit and exactly where the
glass isolation bead is located on the heating element. The thermocouple
converter used in the 540B is selected to have a very low reversal error,
but always will have some error. The error is fixed so you can approximate
a DC measurement once you have characterized the particular converter. I
can't remember now but I think there can be up to around .05% reversal
errors on some converters, while the ones selected for the 540B are under
.01%.
Read the FLUKE "Calibration: Philosophy in Practice" for further
information.
Bill
----- Original Message -----
From: "Dave M" dgminala@mediacombb.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 11:38 AM
Subject: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's
Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer. Leaving the reference voltage setting alone, a DC voltage
input into the unit, and the DC source is adjusted for a null on the
galvanometer. At that point, the AC voltage source is equal to that of
DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you have
have a converter for each range of voltages that you need to measure. The
thermal converters used with this type of transfer standard isn't great
MHz or so typical), but their accuracy far surpasses that of the
type sensors.
There are other brands and models of thermal transfer standards, but I
a Fluke model 540 and a few thermal converters. That's why I referred you
to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two AC+DC
TRMS 7,5 digit meters but the last calibration was 2 years ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for 1V,
10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
and follow the instructions there.
There is one more thing that enters into this discussion and that is
"reversal errors" on the DC. The complicates the transfer somewhat. AC is
always going "+" and "-". DC is in one direction so you have to then
reverse the voltage to the Thermal Transfer Standard and then take the
average of the two readings. That is why there is a "Reversal" switch on
the 540B. When you are using a fixed voltage High Frequency Thermal
Converter you need an external DC reversal switch in addition to other
equipment. You also need an AC/DC transfer switch so that you don't have to
disconnect the AC source and then hook up the DC source manually. See the
540B again.
All Thermal Transfer Standards have some "reversal" error. This is
controlled by the internal construction of the unit and exactly where the
glass isolation bead is located on the heating element. The thermocouple
converter used in the 540B is selected to have a very low reversal error,
but always will have some error. The error is fixed so you can approximate
a DC measurement once you have characterized the particular converter. I
can't remember now but I think there can be up to around .05% reversal
errors on some converters, while the ones selected for the 540B are under
.01%.
Read the FLUKE "Calibration: Philosophy in Practice" for further
information.
Bill
----- Original Message -----
From: "Dave M" <dgminala@mediacombb.net>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Monday, August 25, 2014 11:38 AM
Subject: Re: [volt-nuts] AC calibration
> Well, you sort of answered your own question. The equipment is called a
> Thermal Transfer Standard, but instead of thermistors, it uses a
> thermocouple. Look at the manual for the Fluke 540B
> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's
done.
> Basically, the AC source is input into the transfer standard, and the
> standard's internal reference voltage is adjusted for a null on the
> galvanometer. Leaving the reference voltage setting alone, a DC voltage
is
> input into the unit, and the DC source is adjusted for a null on the
> galvanometer. At that point, the AC voltage source is equal to that of
the
> DC voltage source.
>
> Ther are thermocouple-type thermal converters used for RF voltage
> measurements with the transfer standard. They aren't cheap, and you have
to
> have a converter for each range of voltages that you need to measure. The
> thermal converters used with this type of transfer standard isn't great
(50
> MHz or so typical), but their accuracy far surpasses that of the
thermistor
> type sensors.
>
> There are other brands and models of thermal transfer standards, but I
have
> a Fluke model 540 and a few thermal converters. That's why I referred you
> to the manual for it.
>
> Cheers,
> Dave M
>
>
> pa4tim@gmail.com wrote:
> > Is there a way to link an AC voltage to a DC source for compare. I
> > can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
> > against standardcells. But for AC I can not do that. I have two AC+DC
> > TRMS 7,5 digit meters but the last calibration was 2 years ago.
> >
> > My idea is in theory simple. It is based on the thermal converters
> > used in RF powermeters. Two resistors, two high resolution
> > temperature meters. AC on the first en DC on the second. If both are
> > the same temperature the AC voltage is the same as the DC voltage.
> > But I'm sure some people here have done this in the past. I would
> > like to use it for 50 to 100 kHz (or less) and something like for 1V,
> > 10V and 100V (and use several resistors/heaters.)
> >
> > Or mabey there is an other way to convert AC (for RF it can be done
> > with lightbubs but I never tryed that) I do not mind if it is slow
> > etc, I like this sort of experiments. You can learn a lot from it.
> >
> > Fred, pa4tim
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
D
Don@True-Cal
Tue, Aug 26, 2014 1:54 AM
Why?
Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
Silver or Gold plating on the terminal or wire will introduce the
undesirable dissimilar metal properties, both at the plating junction
and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
--
Mike
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Why?
Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
> Silver or Gold plating on the terminal or wire will introduce the
> undesirable dissimilar metal properties, both at the plating junction
> and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
--
Mike
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
MS
Mike S
Tue, Aug 26, 2014 4:07 AM
On 8/25/2014 9:54 PM, Don@True-Cal wrote:
Why?
Let me count the ways.You can never count on any Seebeck voltage to
be immediately offset, there are far too many variables.
A set of
5440A-7002 (banana plug) cables comes with this calibrator
(5440A-7003 spade lugs for 5720A)
For those leads, Fluke says they have the advantage of "minimizing the
emf caused by temperature difference between the ends of the leads." I
don't have much argument with that - if there's a temperature difference
between the ends, you won't have an equal thermocouple offset, so
minimizing such effects makes sense.
But, since those leads use pure copper wiring ("Belden 8262 or 8719 or
equivalent"), how does having only the contact itself made of tellurium
copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas
tight crimped connections? Does the tinning on that copper cable have an
effect? What's the Seebeck coefficient for a copper to tellurium copper
connection?
You said that "Tellurium Copper is usually not used for a device's
terminal posts but used as the lead wire..." But, Fluke does exactly the
opposite - those Fluke leads use TC connectors and are characterized
"while engaged in a five-way binding post of Tellurium Copper Alloy 145,
half hard." Since they use pure copper cabling, I'd think they'd work
even better with a pure copper binding post, since the offsetting
thermocouples would be in close proximity, and not at different ends of
the cable.
You earlier said "Any type of Silver or Gold plating on the terminal or
wire will introduce the undesirable dissimilar metal properties, both
at the plating junction and at the plating metal to DUT terminal." Now
you emphasize that by saying "If someone used a set of gold-plated
interconnects for this procedure, they would be laughed out of the lab
and the calibrator would be useless until recalibrated properly..."
But, the specs for the Fluke 5440A-7003 which you say are appropriate
for the 5720A state: "Connector materials: Copper space lug with gold
plating over silver plating." (Both platings!!) Is someone to be laughed
out of the lab for using the cables they're supposed to use? The
5440A-7002 specs aren't clear, but if they're not plated, are you
required to clean them of copper oxide for every use? It seems that
would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while
a copper-copper oxide one is ~1000 uV/C.
You seem to be stating one thing, but then giving evidence which
contradicts your claim. It seems that Fluke uses TC where there's
thermal mass, and pure copper where there's little, and uses plating in
addition.
Please clarify, this isn't making sense to me.
On 8/25/2014 9:54 PM, Don@True-Cal wrote:
> Why?
>
> Let me count the ways.You can never count on any Seebeck voltage to
> be immediately offset, there are far too many variables.
...
> A set of
> 5440A-7002 (banana plug) cables comes with this calibrator
> (5440A-7003 spade lugs for 5720A)
For those leads, Fluke says they have the advantage of "minimizing the
emf caused by temperature difference between the ends of the leads." I
don't have much argument with that - if there's a temperature difference
between the ends, you won't have an equal thermocouple offset, so
minimizing such effects makes sense.
But, since those leads use pure copper wiring ("Belden 8262 or 8719 or
equivalent"), how does having only the contact itself made of tellurium
copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas
tight crimped connections? Does the tinning on that copper cable have an
effect? What's the Seebeck coefficient for a copper to tellurium copper
connection?
You said that "Tellurium Copper is usually not used for a device's
terminal posts but used as the lead wire..." But, Fluke does exactly the
opposite - those Fluke leads use TC connectors and are characterized
"while engaged in a five-way binding post of Tellurium Copper Alloy 145,
half hard." Since they use pure copper cabling, I'd think they'd work
even better with a pure copper binding post, since the offsetting
thermocouples would be in close proximity, and not at different ends of
the cable.
You earlier said "Any type of Silver or Gold plating on the terminal or
wire will introduce the undesirable dissimilar metal properties, both
at the plating junction and at the plating metal to DUT terminal." Now
you emphasize that by saying "If someone used a set of gold-plated
interconnects for this procedure, they would be laughed out of the lab
and the calibrator would be useless until recalibrated properly..."
But, the specs for the Fluke 5440A-7003 which you say are appropriate
for the 5720A state: "Connector materials: Copper space lug with gold
plating over silver plating." (Both platings!!) Is someone to be laughed
out of the lab for using the cables they're supposed to use? The
5440A-7002 specs aren't clear, but if they're not plated, are you
required to clean them of copper oxide for every use? It seems that
would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while
a copper-copper oxide one is ~1000 uV/C.
You seem to be stating one thing, but then giving evidence which
contradicts your claim. It seems that Fluke uses TC where there's
thermal mass, and pure copper where there's little, and uses plating in
addition.
Please clarify, this isn't making sense to me.
RE
Randy Evans
Tue, Aug 26, 2014 4:23 AM
Is there any way to tell when the function key routine is complete? In the
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long while
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and the
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
as
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Is there any way to tell when the function key routine is complete? In the
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long while
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
wrote:
> Bill,
>
> I am trying to figure out the MATH function without much success. I input
> the sequence you said (I looked up the instructions to understand what you
> did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
> it shows up on the display when I input BLUE F1. I hit ENTER and it takes
> the 40 measurements and the MATH symbol shows on the display during the
> measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> get a MATH ERR symbol on the display. I tried it a couple of times and the
> same result so I am doing something wrong. Is there a better source for
> explaining how to do front panel masurements than the User Guide, which
> seems oriented at programming automatic rather than manual measurements.
>
> Randy
>
>
>
> On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
>
>> Randy:
>>
>> The MATH function is accessible from the keypad. I don't have an IEEE
>> interface right now that works. You can also program the numeric keypad
>> keys to have preprogrammed functions. DEFKEY
>>
>> I have made my own "low thermal" measurement leads from Pomona #4892
>> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
>> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
>> to do custom cables with 16 ga. bare copper wire that I will twist and
>> then
>> put a braided shield over it. I simply cannot find what I want so I will
>> build my own cable. I have done something like this before and it worked
>> fine. When I get a "round toit".
>>
>> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
>> used
>> in the past to verify my homemade "low thermal" leads as described above.
>> Frankly I cannot see any difference between using the 1756 cables and my
>> homemade cables once I give them a few minutes for the thermals to go
>> away.
>> As far as I can tell and measure the differences, if any, are below 0.1
>> ppm
>> at 10 volts.
>>
>> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
>> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
>> as
>> the instability of the readings it is hard to determine which is causing
>> the
>> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
>> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
>> set
>> the MATH to "Statistics" (store high reading/low reading/ and mean of the
>> readings) in the registers, the number of readings to "40", the trigger to
>> "hold" (which keeps the meter from triggering until I press "ENTER" and
>> then
>> trigger the sequence of 40 readings when I push the "ENTER" button. You
>> can
>> do all of this manually from the keypads but since I use this sequence a
>> lot
>> I have preprogrammed it. This is after I set digits to "8" and PLC to
>> 100.
>> Once those 40 readings are finished then you can access the various MATH
>> statistic registers, using the menu, by entering "MATH" and then a 2 for
>> low, a 4 for mean, and 13 for high. Of course you could do all of this
>> through the IEEE also. The 3458A has a very rich set of measurement
>> commands. I am still learning all of them. It depends upon what I am
>> trying to accomplish.
>>
>> Since the 1.018 and 1.0 volt outputs are passive and derived from
>> resistive dividers from the 10 volt, I don't see how they could contribute
>> to the varying readings you are measuring. I think I would put a short on
>> the input of the 3458A and manually set the range to 1 volt and then
>> observe
>> the variations that way without the 732A involved. When I do this I see a
>> variation from low reading to high reading of 0.125 uVolts and then
>> another
>> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
>> side
>> of the measurment terminals, GUARD connected doesn't seem to affect the
>> readings. So that is the base noise of the 3458A without the 732A,
>> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
>> got
>> a variation of 0.159 uVolts using the same 40 reading method above. I
>> would
>> use this to determine where your problem might exist. Just having the
>> meter
>> input shorted will point you in the right direction. Meter, cables or
>> 732A.
>>
>> Sorry for the long dissertation. Friends get mad at me for being so
>> detailed sometimes.
>>
>> Bill
>>
>> ----- Original Message -----
>> From: "Randy Evans" <randyevans2688@gmail.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Sent: Sunday, August 24, 2014 7:22 AM
>> Subject: Re: [volt-nuts] 732A and Prologix received
>>
>>
>> > Bill,
>> >
>> > I have convinced myself that the problem I an seeing is due to thermals.
>> > If I move the cables (with gold-plated banana plugs) using a small
>> towel
>> > rather than letting my hand touch the plugs, it is much more stable.
>> If I
>> > then hold the banana plug with my hand after the reading has stabilized,
>> > the reading drifts rapidly upward. I am trying to check the stability
>> of
>> > the reading but I haven't figured out the MATH function yet. I assume
>> this
>> > is a programmed function using GPIB only?
>> >
>> > The stability I am seeing by manually recording readings (using NLPC of
>> 100
>> > and 1000) is much greater than what you are measuring on your system.
>> Not
>> > sure how to ascertain if it's the 3458A or the 732. The value of the
>> > readings are very different between the two - the 3458 reads about 50 uV
>> > high on the 10 V output and about 12 uV low on the 1V output. Rather
>> large
>> > differences (this is after an ACAL). I need to find some better cables
>> to
>> > make sure the errors are not due to thermals again.
>> >
>> > Randy
>> >
>> >
>> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
>> >
>> > > Randy:
>> > >
>> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
>> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
>> > > others
>> > > that are in this size and package. Order from one of the usual
>> electronics
>> > > distributors like Allied, Mouser, Digikey. This is a very common
>> battery
>> > > as
>> > > it is used in a lot of "EXIT" signs so they are lighted when the power
>> goes
>> > > out. I don't see how the 12V 7AH will fit as they are too large. I
>> guess
>> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt
>> 4
>> AH
>> > > but the terminals are in the wrong place so you will have to "nibble"
>> out
>> > > the aluminum plate that holds them in the 732A battery pack. You have
>> to
>> > > be
>> > > careful if you use the 12v 5AH as you will have 4 extra battery
>> connection
>> > > leads to deal with and connect correctly. I would stick with the 6V
>> 4AH.
>> > > New batteries will last around 12 to 14 hours before the "CAL" light
>> goes
>> > > out when AC power is not applied. So shipping to Cal Lab can be a
>> problem
>> > > if it is a distance away, or you have to use a shipper like UPS or
>> FEDEX
>> > > and
>> > > you ship the night before and then use their "Morning delivery" and
>> the
>> Cal
>> > > Lab is expecting your 732A. Same on the way back to you. Of course
>> you
>> > > could always strap another battery on the 732A and hook it up to the
>> "ext
>> > > power" plug to last longer. I have seen it done. The issue is to get
>> the
>> > > Cal Lab to charge the extra battery before they ship the 732A back to
>> you.
>> > >
>> > > When you remove and work on the battery pack always have the AC
>> power
>> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
>> supply
>> > > is working.
>> > >
>> > > The "CAL" light is to indicate that power has not been lost to the
>> > > Reference Amp or other associated circuits. When the raw supply
>> (battery)
>> > > voltage drops below about 21 volts the "CAL" light will go out. Below
>> that
>> > > voltage the heater circuits will not work correctly and the 18.6 volt
>> > > regulated supply will not regulate. The requirement is that the
>> Reference
>> > > Amp be kept "alive" at all times to maintain the output voltage that
>> was
>> > > measured at the time of the most recent Calibration or Certification.
>> When
>> > > the semiconductor junctions are unbiased and cool off when power is
>> lost,
>> > > and then power is restored the result will be a different 10 volts
>> than
>> > > before the power failure. My experience is that after all of the
>> years
>> > > that
>> > > these units have been powered up, this won't happen and when power is
>> lost
>> > > and then restored, even months later, the 732A will come back to
>> almost
>> > > exactly the same 10 volts as when they lost power, usually with in 0.2
>> PPM
>> > > after 24 hours of "warm up".
>> > >
>> > > What type of hookup leads are you using when measuring the 1 volt
>> > > output? What is the PLC set to? I always use 100 PLC to measure
>> this.
>> If
>> > > you don't have "low thermal" connection leads you can experience uV
>> changes
>> > > for a minute or more after plugging in the leads due to the "thermals"
>> > > generated because of the difference in temperature between the banana
>> jacks
>> > > on the 732A and the banana plugs of the leads. I have found that even
>> just
>> > > plugging in the lead will generate a thermal difference because of
>> > > difference of temps and some heating due to the physical act of just
>> > > inserting the plug because of friction between the jack and plug (my
>> theory
>> > > at any rate). You have to allow at least a minute or more before
>> being
>> > > able
>> > > to make a measurement after plugging in the leads. I just measured
>> the
>> > > variation of the 1 volt output of my 732A and using my 3458A and I
>> got a
>> > > total difference of 0.159 uV over 40 measurements using 100 PLC on
>> the
>> 1
>> > > volt range of the 3458A. Using the MATH function and all of the data
>> you
>> > > can collect. That was after waiting for several minutes after
>> plugging
>> in
>> > > the leads.
>> > >
>> > > I hope all of this helps.
>> > >
>> > > Bill
>> > >
>> > >
>> > >
>> > > ----- Original Message -----
>> > > From: "Randy Evans" <randyevans2688@gmail.com>
>> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> > > Sent: Friday, August 22, 2014 9:03 PM
>> > > Subject: Re: [volt-nuts] 732A and Prologix received
>> > >
>> > >
>> > > > Todd,
>> > > >
>> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
>> that
>> I
>> > > > keep topped off and they have very low current draw (~2 to 3 mA at
>> 13.5
>> > > > VDC) when charged and at their float voltage, so I am pretty sure
>> they
>> > > are
>> > > > in good condition. I will look at getting those in the units after
>> I
>> > > > ascertain the condition of the 732.
>> > > >
>> > > > So now I have a what appears to be a functioning 3458A and a 732A
>> but
>> > > they
>> > > > slightly disagree. I am like the man with two watches that disagree
>> on
>> > > the
>> > > > time - which is correct? For the moment, i am only concerned with
>> > > > stability. The need for absolute accuracy will come later.
>> > > >
>> > > > Randy
>> > > >
>> > > >
>> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
>> >
>> > > wrote:
>> > > >
>> > > > > Randy,
>> > > > >
>> > > > > You have two possible choices. It can be configured with 4 x 6v
>> 4Ah
>> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
>> has
>> > > > > modified the battery pack already. A couple of mine needed a
>> nibbler
>> > > tool
>> > > > > to remove enough of the aluminum cover that fits over the tops of
>> the
>> > > > > batteries. The original cover will short out to the battery tabs
>> > > regardless
>> > > > > of the battery configuration if this is not done.
>> > > > >
>> > > > > You can find larger capacity batteries that will give you slightly
>> more
>> > > > > battery life. I lost a couple sets of mail-order batteries after a
>> few
>> > > > > extended outages. I would recommend going with locally bought
>> batteries
>> > > > > instead of the cheaper mail order. My local Batteries Plus will
>> > > typically
>> > > > > have some warranty if I remember correctly. Moving forward I will
>> only
>> > > use
>> > > > > 2 12v batteries and pre-charge them on a battery charger to
>> equalize
>> > > them
>> > > > > before putting them in the 732A. I think the cheap batteries did
>> not
>> > > > > discharge equally, and would not recover when power was applied.
>> > > > >
>> > > > > Inspect the back plane for damaged traces and look at the
>> capacitors. I
>> > > had
>> > > > > a few that looked questionable. So far, I have replaced all the
>> big
>> > > caps
>> > > on
>> > > > > the pre-regulator and regulator boards. My feeling is that once
>> these
>> > > go
>> > > > > online, they should run as long as possible between repairs.
>> > > > >
>> > > > > The battery charger circuit may need adjusting. I tweaked mine and
>> it
>> > > > > seemed to work fine.
>> > > > >
>> > > > > Todd
>> > > > >
>> > > > >
>> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
>> > > randyevans2688@gmail.com>
>> > > > > wrote:
>> > > > >
>> > > > > > I received my Fluke 732A today. Just powered it up but it needs
>> new
>> > > > > > batteries. Any suggestions for sources (I haven't opened up the
>> unit
>> > > > > yet -
>> > > > > > I want to make sure it works before doing that). Also received
>> the
>> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
>> data
>> > > > > dumper
>> > > > > > program to get the CAL data from my 3458A. Should be a busy
>> weekend.
>> > > > > >
>> > > > > > Randy
>> > > > > > _______________________________________________
>> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > > > To unsubscribe, go to
>> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > > > and follow the instructions there.
>> > > > > >
>> > > > > _______________________________________________
>> > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > > To unsubscribe, go to
>> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > > and follow the instructions there.
>> > > > >
>> > > > _______________________________________________
>> > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > To unsubscribe, go to
>> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > and follow the instructions there.
>> > >
>> > > _______________________________________________
>> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > To unsubscribe, go to
>> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > and follow the instructions there.
>> > >
>> > _______________________________________________
>> > volt-nuts mailing list -- volt-nuts@febo.com
>> > To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > and follow the instructions there.
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
>
CB
Charles Black
Tue, Aug 26, 2014 4:28 AM
My post on this thread earlier today seems to be lost (maybe the system
change) so I am re-posting some of it (corrected) since it's apropos to
the discussion.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire. This is very convenient since it is so
easy to duplicate in the field and makes simple inexpensive test leads
best for high precision measurements. In order for the 3458A to make
full accuracy measurements (8 digit) NPLC must be set to 1000 (according
to the User's Guide). If you use a lower NPLC value there is a table in
the User's Guide that can be used to determine how accurate your
measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv ± 10nV Equilibration time 5
minutes. 14 gauge per Calibration Manual.
"U" heavy wire -0.00021mv ± 10nV Equil. time 5 minutes.
Used the through holes in the Input banana posts only.
Copper wire -0.00019mv ± 10nV NAPA PVC covered
automobile wire at same contact points as CAL shunt
Standard Ground Plate -0.00040mv ± 10nV Equil. time 5 minutes. Gold
plated ground plate from my Datron 4910
Copper wire -0.00019mv ± 10nv Equil. time 2
seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv ± 10nV Equil. time 2 seconds.
M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv ± 30nV Equil. time 20
minutes. My best "Perfect" gold plated plugs with copper wire.
Note that all my test shorts equilibrate to virtualy zero volts. There
is almost no Seebeck voltage for any of my test shorts as long as the
system is allowed to equilibrate. Equilibation times varied between less
than 2 seconds up 20 minutes with longer times due to excessive thermal
mass. Excessive thermal mass also caused some voltage instability.
Charlie
On 8/25/2014 6:54 PM, Don@True-Cal wrote:
Why?
Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
Silver or Gold plating on the terminal or wire will introduce the
undesirable dissimilar metal properties, both at the plating junction
and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
--
Mike
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
My post on this thread earlier today seems to be lost (maybe the system
change) so I am re-posting some of it (corrected) since it's apropos to
the discussion.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire. This is very convenient since it is so
easy to duplicate in the field and makes simple inexpensive test leads
best for high precision measurements. In order for the 3458A to make
full accuracy measurements (8 digit) NPLC must be set to 1000 (according
to the User's Guide). If you use a lower NPLC value there is a table in
the User's Guide that can be used to determine how accurate your
measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv ± 10nV Equilibration time 5
minutes. 14 gauge per Calibration Manual.
"U" heavy wire -0.00021mv ± 10nV Equil. time 5 minutes.
Used the through holes in the Input banana posts only.
Copper wire -0.00019mv ± 10nV NAPA PVC covered
automobile wire at same contact points as CAL shunt
Standard Ground Plate -0.00040mv ± 10nV Equil. time 5 minutes. Gold
plated ground plate from my Datron 4910
Copper wire -0.00019mv ± 10nv Equil. time 2
seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv ± 10nV Equil. time 2 seconds.
M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv ± 30nV Equil. time 20
minutes. My best "Perfect" gold plated plugs with copper wire.
Note that all my test shorts equilibrate to virtualy zero volts. There
is almost no Seebeck voltage for any of my test shorts as long as the
system is allowed to equilibrate. Equilibation times varied between less
than 2 seconds up 20 minutes with longer times due to excessive thermal
mass. Excessive thermal mass also caused some voltage instability.
Charlie
On 8/25/2014 6:54 PM, Don@True-Cal wrote:
> Why?
>
> Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
>
> BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
>
> This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
>
> Don Johnson
>
> -----Original Message-----
> From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
> Sent: Monday, August 25, 2014 11:03 AM
> To: volt-nuts@febo.com
> Subject: Re: [volt-nuts] 732A and Prologix received
>
> On 8/25/2014 11:02 AM, Don@True-Cal wrote:
>> Silver or Gold plating on the terminal or wire will introduce the
>> undesirable dissimilar metal properties, both at the plating junction
>> and at the plating metal to DUT terminal.
> Why?
>
> Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
>
> A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
> shortly after plugging in a banana plug warmed by body heat), they simply cancel.
>
> Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
>
> --
> Mike
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
>
BG
Bill Gold
Tue, Aug 26, 2014 5:01 AM
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front panel
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand what
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy at
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the trigger
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this sequence
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a short
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too large.
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is to
Cal Lab to charge the extra battery before they ship the 732A back
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost to
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened up
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front panel
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
> Is there any way to tell when the function key routine is complete? In
the
> case of taking multiple readings using the DEFKEY and MATH function, I
> don't see any indication when the routine is complete. In one particular
> case, I am taking a 100 readings with NLPC set for 1000 so its a long
while
> before it's complete, but i have to guess when it's done.
>
> Thanks,
>
> Randy
>
>
> On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > Bill,
> >
> > I am trying to figure out the MATH function without much success. I
input
> > the sequence you said (I looked up the instructions to understand what
you
> > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
and
> > it shows up on the display when I input BLUE F1. I hit ENTER and it
takes
> > the 40 measurements and the MATH symbol shows on the display during the
> > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> > get a MATH ERR symbol on the display. I tried it a couple of times and
the
> > same result so I am doing something wrong. Is there a better source for
> > explaining how to do front panel masurements than the User Guide, which
> > seems oriented at programming automatic rather than manual measurements.
> >
> > Randy
> >
> >
> >
> > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> >
> >> Randy:
> >>
> >> The MATH function is accessible from the keypad. I don't have an
IEEE
> >> interface right now that works. You can also program the numeric
keypad
> >> keys to have preprogrammed functions. DEFKEY
> >>
> >> I have made my own "low thermal" measurement leads from Pomona
#4892
> >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at
the
> >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
plans
> >> to do custom cables with 16 ga. bare copper wire that I will twist and
> >> then
> >> put a braided shield over it. I simply cannot find what I want so I
will
> >> build my own cable. I have done something like this before and it
worked
> >> fine. When I get a "round toit".
> >>
> >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> >> used
> >> in the past to verify my homemade "low thermal" leads as described
above.
> >> Frankly I cannot see any difference between using the 1756 cables and
my
> >> homemade cables once I give them a few minutes for the thermals to go
> >> away.
> >> As far as I can tell and measure the differences, if any, are below 0.1
> >> ppm
> >> at 10 volts.
> >>
> >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
all
> >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
far
> >> as
> >> the instability of the readings it is hard to determine which is
causing
> >> the
> >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
is
> >> set
> >> the MATH to "Statistics" (store high reading/low reading/ and mean of
the
> >> readings) in the registers, the number of readings to "40", the trigger
to
> >> "hold" (which keeps the meter from triggering until I press "ENTER" and
> >> then
> >> trigger the sequence of 40 readings when I push the "ENTER" button.
You
> >> can
> >> do all of this manually from the keypads but since I use this sequence
a
> >> lot
> >> I have preprogrammed it. This is after I set digits to "8" and PLC to
> >> 100.
> >> Once those 40 readings are finished then you can access the various
MATH
> >> statistic registers, using the menu, by entering "MATH" and then a 2
for
> >> low, a 4 for mean, and 13 for high. Of course you could do all of this
> >> through the IEEE also. The 3458A has a very rich set of measurement
> >> commands. I am still learning all of them. It depends upon what I am
> >> trying to accomplish.
> >>
> >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> >> resistive dividers from the 10 volt, I don't see how they could
contribute
> >> to the varying readings you are measuring. I think I would put a short
on
> >> the input of the 3458A and manually set the range to 1 volt and then
> >> observe
> >> the variations that way without the 732A involved. When I do this I
see a
> >> variation from low reading to high reading of 0.125 uVolts and then
> >> another
> >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
> >> side
> >> of the measurment terminals, GUARD connected doesn't seem to affect the
> >> readings. So that is the base noise of the 3458A without the 732A,
> >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
> >> got
> >> a variation of 0.159 uVolts using the same 40 reading method above. I
> >> would
> >> use this to determine where your problem might exist. Just having the
> >> meter
> >> input shorted will point you in the right direction. Meter, cables or
> >> 732A.
> >>
> >> Sorry for the long dissertation. Friends get mad at me for being
so
> >> detailed sometimes.
> >>
> >> Bill
> >>
> >> ----- Original Message -----
> >> From: "Randy Evans" <randyevans2688@gmail.com>
> >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> Sent: Sunday, August 24, 2014 7:22 AM
> >> Subject: Re: [volt-nuts] 732A and Prologix received
> >>
> >>
> >> > Bill,
> >> >
> >> > I have convinced myself that the problem I an seeing is due to
thermals.
> >> > If I move the cables (with gold-plated banana plugs) using a small
> >> towel
> >> > rather than letting my hand touch the plugs, it is much more stable.
> >> If I
> >> > then hold the banana plug with my hand after the reading has
stabilized,
> >> > the reading drifts rapidly upward. I am trying to check the
stability
> >> of
> >> > the reading but I haven't figured out the MATH function yet. I
assume
> >> this
> >> > is a programmed function using GPIB only?
> >> >
> >> > The stability I am seeing by manually recording readings (using NLPC
of
> >> 100
> >> > and 1000) is much greater than what you are measuring on your system.
> >> Not
> >> > sure how to ascertain if it's the 3458A or the 732. The value of the
> >> > readings are very different between the two - the 3458 reads about 50
uV
> >> > high on the 10 V output and about 12 uV low on the 1V output. Rather
> >> large
> >> > differences (this is after an ACAL). I need to find some better
cables
> >> to
> >> > make sure the errors are not due to thermals again.
> >> >
> >> > Randy
> >> >
> >> >
> >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
wrote:
> >> >
> >> > > Randy:
> >> > >
> >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
fit
> >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
and
> >> > > others
> >> > > that are in this size and package. Order from one of the usual
> >> electronics
> >> > > distributors like Allied, Mouser, Digikey. This is a very common
> >> battery
> >> > > as
> >> > > it is used in a lot of "EXIT" signs so they are lighted when the
power
> >> goes
> >> > > out. I don't see how the 12V 7AH will fit as they are too large.
I
> >> guess
> >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
volt
> >> 4
> >> AH
> >> > > but the terminals are in the wrong place so you will have to
"nibble"
> >> out
> >> > > the aluminum plate that holds them in the 732A battery pack. You
have
> >> to
> >> > > be
> >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> >> connection
> >> > > leads to deal with and connect correctly. I would stick with the
6V
> >> 4AH.
> >> > > New batteries will last around 12 to 14 hours before the "CAL"
light
> >> goes
> >> > > out when AC power is not applied. So shipping to Cal Lab can be a
> >> problem
> >> > > if it is a distance away, or you have to use a shipper like UPS or
> >> FEDEX
> >> > > and
> >> > > you ship the night before and then use their "Morning delivery" and
> >> the
> >> Cal
> >> > > Lab is expecting your 732A. Same on the way back to you. Of
course
> >> you
> >> > > could always strap another battery on the 732A and hook it up to
the
> >> "ext
> >> > > power" plug to last longer. I have seen it done. The issue is to
get
> >> the
> >> > > Cal Lab to charge the extra battery before they ship the 732A back
to
> >> you.
> >> > >
> >> > > When you remove and work on the battery pack always have the AC
> >> power
> >> > > plugged in. The "CAL" led will stay on because the 18.6 v
regulated
> >> supply
> >> > > is working.
> >> > >
> >> > > The "CAL" light is to indicate that power has not been lost to
the
> >> > > Reference Amp or other associated circuits. When the raw supply
> >> (battery)
> >> > > voltage drops below about 21 volts the "CAL" light will go out.
Below
> >> that
> >> > > voltage the heater circuits will not work correctly and the 18.6
volt
> >> > > regulated supply will not regulate. The requirement is that the
> >> Reference
> >> > > Amp be kept "alive" at all times to maintain the output voltage
that
> >> was
> >> > > measured at the time of the most recent Calibration or
Certification.
> >> When
> >> > > the semiconductor junctions are unbiased and cool off when power is
> >> lost,
> >> > > and then power is restored the result will be a different 10 volts
> >> than
> >> > > before the power failure. My experience is that after all of the
> >> years
> >> > > that
> >> > > these units have been powered up, this won't happen and when power
is
> >> lost
> >> > > and then restored, even months later, the 732A will come back to
> >> almost
> >> > > exactly the same 10 volts as when they lost power, usually with in
0.2
> >> PPM
> >> > > after 24 hours of "warm up".
> >> > >
> >> > > What type of hookup leads are you using when measuring the 1
volt
> >> > > output? What is the PLC set to? I always use 100 PLC to measure
> >> this.
> >> If
> >> > > you don't have "low thermal" connection leads you can experience uV
> >> changes
> >> > > for a minute or more after plugging in the leads due to the
"thermals"
> >> > > generated because of the difference in temperature between the
banana
> >> jacks
> >> > > on the 732A and the banana plugs of the leads. I have found that
even
> >> just
> >> > > plugging in the lead will generate a thermal difference because of
> >> > > difference of temps and some heating due to the physical act of
just
> >> > > inserting the plug because of friction between the jack and plug
(my
> >> theory
> >> > > at any rate). You have to allow at least a minute or more before
> >> being
> >> > > able
> >> > > to make a measurement after plugging in the leads. I just measured
> >> the
> >> > > variation of the 1 volt output of my 732A and using my 3458A and I
> >> got a
> >> > > total difference of 0.159 uV over 40 measurements using 100 PLC on
> >> the
> >> 1
> >> > > volt range of the 3458A. Using the MATH function and all of the
data
> >> you
> >> > > can collect. That was after waiting for several minutes after
> >> plugging
> >> in
> >> > > the leads.
> >> > >
> >> > > I hope all of this helps.
> >> > >
> >> > > Bill
> >> > >
> >> > >
> >> > >
> >> > > ----- Original Message -----
> >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> >> > > To: "Discussion of precise voltage measurement"
<volt-nuts@febo.com>
> >> > > Sent: Friday, August 22, 2014 9:03 PM
> >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >
> >> > >
> >> > > > Todd,
> >> > > >
> >> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> >> that
> >> I
> >> > > > keep topped off and they have very low current draw (~2 to 3 mA
at
> >> 13.5
> >> > > > VDC) when charged and at their float voltage, so I am pretty sure
> >> they
> >> > > are
> >> > > > in good condition. I will look at getting those in the units
after
> >> I
> >> > > > ascertain the condition of the 732.
> >> > > >
> >> > > > So now I have a what appears to be a functioning 3458A and a 732A
> >> but
> >> > > they
> >> > > > slightly disagree. I am like the man with two watches that
disagree
> >> on
> >> > > the
> >> > > > time - which is correct? For the moment, i am only concerned
with
> >> > > > stability. The need for absolute accuracy will come later.
> >> > > >
> >> > > > Randy
> >> > > >
> >> > > >
> >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
<tmicallef@gmail.com
> >> >
> >> > > wrote:
> >> > > >
> >> > > > > Randy,
> >> > > > >
> >> > > > > You have two possible choices. It can be configured with 4 x 6v
> >> 4Ah
> >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
owner
> >> has
> >> > > > > modified the battery pack already. A couple of mine needed a
> >> nibbler
> >> > > tool
> >> > > > > to remove enough of the aluminum cover that fits over the tops
of
> >> the
> >> > > > > batteries. The original cover will short out to the battery
tabs
> >> > > regardless
> >> > > > > of the battery configuration if this is not done.
> >> > > > >
> >> > > > > You can find larger capacity batteries that will give you
slightly
> >> more
> >> > > > > battery life. I lost a couple sets of mail-order batteries
after a
> >> few
> >> > > > > extended outages. I would recommend going with locally bought
> >> batteries
> >> > > > > instead of the cheaper mail order. My local Batteries Plus will
> >> > > typically
> >> > > > > have some warranty if I remember correctly. Moving forward I
will
> >> only
> >> > > use
> >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> >> equalize
> >> > > them
> >> > > > > before putting them in the 732A. I think the cheap batteries
did
> >> not
> >> > > > > discharge equally, and would not recover when power was
applied.
> >> > > > >
> >> > > > > Inspect the back plane for damaged traces and look at the
> >> capacitors. I
> >> > > had
> >> > > > > a few that looked questionable. So far, I have replaced all the
> >> big
> >> > > caps
> >> > > on
> >> > > > > the pre-regulator and regulator boards. My feeling is that once
> >> these
> >> > > go
> >> > > > > online, they should run as long as possible between repairs.
> >> > > > >
> >> > > > > The battery charger circuit may need adjusting. I tweaked mine
and
> >> it
> >> > > > > seemed to work fine.
> >> > > > >
> >> > > > > Todd
> >> > > > >
> >> > > > >
> >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> >> > > randyevans2688@gmail.com>
> >> > > > > wrote:
> >> > > > >
> >> > > > > > I received my Fluke 732A today. Just powered it up but it
needs
> >> new
> >> > > > > > batteries. Any suggestions for sources (I haven't opened up
the
> >> unit
> >> > > > > yet -
> >> > > > > > I want to make sure it works before doing that). Also
received
> >> the
> >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
ran
> >> data
> >> > > > > dumper
> >> > > > > > program to get the CAL data from my 3458A. Should be a busy
> >> weekend.
> >> > > > > >
> >> > > > > > Randy
> >> > > > > > _______________________________________________
> >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > > > To unsubscribe, go to
> >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > > > and follow the instructions there.
> >> > > > > >
> >> > > > > _______________________________________________
> >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > > To unsubscribe, go to
> >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > > and follow the instructions there.
> >> > > > >
> >> > > > _______________________________________________
> >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > To unsubscribe, go to
> >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > and follow the instructions there.
> >> > >
> >> > > _______________________________________________
> >> > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > To unsubscribe, go to
> >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > and follow the instructions there.
> >> > >
> >> > _______________________________________________
> >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > and follow the instructions there.
> >>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >>
> >
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
MS
Mike S
Tue, Aug 26, 2014 3:05 PM
After some more research, I think I've answered some of my own questions -
Tellurium copper is used for binding posts, not because it has any
special thermal or EMF mojo, but because it machines much better than
pure copper. And, I suppose, because it sounds like it's extra special.
The Seebeck coefficients (uV/C, relative to Cu) of some relevant materials:
Cu 0.0
Ag .2
Au .5
Yellow brass 1.5
Phosphor bronze 2.0
63/37 solder 3.0
Sn 3.1
Stainless steel 3.1
Beryllium copper 5.0
Fe -12.3
Ni 22.3
Te -49.25
Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
I'd expect that there might be some coefficient between Cu and CuTe
(0.5% Te), but I could find no reference. The relatively large number
for CuBe is interesting, since that's a common material for banana plug
springs, where one might expect the greatest temperature differential to
occur in such a connection (between the thermal masses of the binding
post/jack and the bulk of the banana plug). Heat has to flow a
considerable distance through the springs, very much more than when it
flows through a surface plating.
The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
contacts. It seems there might be an improvement to be had by using the
older style pin plugs, where a solid pin was partially sliced into 4
sections which were then spread apart a bit to create tension. That
could eliminate relatively large thermocouples at a thermal gradient,
and might also be expected to have less thermal resistance, allowing the
connection to settle quicker.
But maybe not - I'm still not clear on how plated conductors behave in
this situation. For a high impedance voltage measurement where almost no
current flows, the gold plating may carry the signal, so there is no
real thermocouple (or more correctly, it's entirely contained within the
connector). But if that's the case, why fool around with special copper
connectors when common brass ones would be easier/cheaper? For current
or resistance, the signal would also flow through the base metal, so
does this have an effect (especially for tinned copper test leads, where
there may be a larger temperature difference between the ends???
Nickle is avoided as a contact material largely because it is subject to
fretting corrosion. Tests done by AMP
(http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show that
a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a short
time due to this, while Ag and Au plated contacts exhibit negligible
changes.
Cu (with Be for better machining) seems to be used as the base material
for jacks/plugs to get thermal EMF cancellation to the wiring on both
sides (i.e. use copper everywhere except where there is a minimal
thermal gradient, like platings).
--
After some more research, I think I've answered some of my own questions -
Tellurium copper is used for binding posts, not because it has any
special thermal or EMF mojo, but because it machines much better than
pure copper. And, I suppose, because it sounds like it's extra special.
The Seebeck coefficients (uV/C, relative to Cu) of some relevant materials:
Cu 0.0
Ag .2
Au .5
Yellow brass 1.5
Phosphor bronze 2.0
63/37 solder 3.0
Sn 3.1
Stainless steel 3.1
Beryllium copper 5.0
Fe -12.3
Ni 22.3
Te -49.25
Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
I'd expect that there might be some coefficient between Cu and CuTe
(0.5% Te), but I could find no reference. The relatively large number
for CuBe is interesting, since that's a common material for banana plug
springs, where one might expect the greatest temperature differential to
occur in such a connection (between the thermal masses of the binding
post/jack and the bulk of the banana plug). Heat has to flow a
considerable distance through the springs, very much more than when it
flows through a surface plating.
The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
contacts. It seems there might be an improvement to be had by using the
older style pin plugs, where a solid pin was partially sliced into 4
sections which were then spread apart a bit to create tension. That
could eliminate relatively large thermocouples at a thermal gradient,
and might also be expected to have less thermal resistance, allowing the
connection to settle quicker.
But maybe not - I'm still not clear on how plated conductors behave in
this situation. For a high impedance voltage measurement where almost no
current flows, the gold plating may carry the signal, so there is no
real thermocouple (or more correctly, it's entirely contained within the
connector). But if that's the case, why fool around with special copper
connectors when common brass ones would be easier/cheaper? For current
or resistance, the signal would also flow through the base metal, so
does this have an effect (especially for tinned copper test leads, where
there may be a larger temperature difference between the ends???
Nickle is avoided as a contact material largely because it is subject to
fretting corrosion. Tests done by AMP
(http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show that
a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a short
time due to this, while Ag and Au plated contacts exhibit negligible
changes.
Cu (with Be for better machining) seems to be used as the base material
for jacks/plugs to get thermal EMF cancellation to the wiring on both
sides (i.e. use copper everywhere except where there is a minimal
thermal gradient, like platings).
--
A
acbern@gmx.de
Tue, Aug 26, 2014 4:56 PM
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Dave M
acbern@gmx.de wrote:
fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be <10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.
adrian
Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer. Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer. At that point,
the AC voltage source is equal to that of the DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you
have to have a converter for each range of voltages that you need to
measure. The thermal converters used with this type of transfer
standard isn't great (50 MHz or so typical), but their accuracy far
surpasses that of the thermistor type sensors.
There are other brands and models of thermal transfer standards, but
I have a Fluke model 540 and a few thermal converters. That's why I
referred you to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two
AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for
1V, 10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.
Adrian
> Gesendet: Montag, 25. August 2014 um 19:32 Uhr
> Von: "Dave M" <dgminala@mediacombb.net>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] AC calibration
>
> Adrian,
> Do you have a link or title for the NIST paper that you mentioned?
>
> Dave M
>
>
> acbern@gmx.de wrote:
> > fred,
> > generally you raise a good point, I had the same issue of calibrating
> > an ac voltage to a high level of accuracy. you need this e.g. to
> > validate the self.cal of a 3458a or other precison stuff like the
> > 8506a0.
> >
> > what i would recommend to do if you want to keep costs down is:
> > in a nutshell, get a thermal converter in the lowest range you need
> > and a second one on range above. build a set of resistor range
> > extenders (rf type with appropriate connectors and housings) to
> > expand the range to where you need to be max. get one of the thermal
> > converter calibrated (the higher one usually, and you need to havr
> > good cal lab, should be <10ppm accuracy) and use it to calibrate the
> > rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
> > thermal converters! at higher frequencies, due to reflections and
> > stray capacitance/inductance influences, the accuracy decreases. the
> > resistor range extenders though, if build up correctly, only have a
> > few ppm impact (there is a paper from nist on that, but this is only
> > typical). you can calibrate all converters to the one you got
> > externally calibrated. do some research in the web, when you do the
> > calibration, you need to determine the so-called constant N. then do
> > an ac, dc+, ac, dc-, ac measurement between the the two and establish
> > the deviation, also establish the error propagation. the end result
> > will be a set of highly precise (low inaccuracies9 thermal converters
> > good enough to calibrate a 3458a an better devices. if you want to
> > spend the money, you could also buy a set of converters/range
> > resistors (with/without a 540), that typically is a few k altogether,
> > while a single device sometimes is available for below 100 bucks. you
> > need to have a stable 7.5 digit nanovoltmeter though for the
> > measurements of the tvcs (34420a or 2182 typically ) and precision
> > (stable) dc and ac sources. but in the end, all you need is a single
> > calibrated thermal converter.
> >
> > adrian
> >
> >
> >
> >> Gesendet: Montag, 25. August 2014 um 18:38 Uhr
> >> Von: "Dave M" <dgminala@mediacombb.net>
> >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> Betreff: Re: [volt-nuts] AC calibration
> >>
> >> Well, you sort of answered your own question. The equipment is
> >> called a Thermal Transfer Standard, but instead of thermistors, it
> >> uses a thermocouple. Look at the manual for the Fluke 540B
> >> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
> >> it's done. Basically, the AC source is input into the transfer
> >> standard, and the standard's internal reference voltage is adjusted
> >> for a null on the galvanometer. Leaving the reference voltage
> >> setting alone, a DC voltage is input into the unit, and the DC
> >> source is adjusted for a null on the galvanometer. At that point,
> >> the AC voltage source is equal to that of the DC voltage source.
> >>
> >> Ther are thermocouple-type thermal converters used for RF voltage
> >> measurements with the transfer standard. They aren't cheap, and you
> >> have to have a converter for each range of voltages that you need to
> >> measure. The thermal converters used with this type of transfer
> >> standard isn't great (50 MHz or so typical), but their accuracy far
> >> surpasses that of the thermistor type sensors.
> >>
> >> There are other brands and models of thermal transfer standards, but
> >> I have a Fluke model 540 and a few thermal converters. That's why I
> >> referred you to the manual for it.
> >>
> >> Cheers,
> >> Dave M
> >>
> >>
> >> pa4tim@gmail.com wrote:
> >>> Is there a way to link an AC voltage to a DC source for compare. I
> >>> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
> >>> against standardcells. But for AC I can not do that. I have two
> >>> AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
> >>> ago.
> >>>
> >>> My idea is in theory simple. It is based on the thermal converters
> >>> used in RF powermeters. Two resistors, two high resolution
> >>> temperature meters. AC on the first en DC on the second. If both are
> >>> the same temperature the AC voltage is the same as the DC voltage.
> >>> But I'm sure some people here have done this in the past. I would
> >>> like to use it for 50 to 100 kHz (or less) and something like for
> >>> 1V, 10V and 100V (and use several resistors/heaters.)
> >>>
> >>> Or mabey there is an other way to convert AC (for RF it can be done
> >>> with lightbubs but I never tryed that) I do not mind if it is slow
> >>> etc, I like this sort of experiments. You can learn a lot from it.
> >>>
> >>> Fred, pa4tim
> >>
> >>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
> >> the instructions there.
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> When the people fear the government, there is tyranny. When the
> government fears the people, there is liberty -- Thomas Jefferson
>
>
> Dave M
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
PK
Poul-Henning Kamp
Tue, Aug 26, 2014 7:42 PM
In message <trinity-9076c25b-a0a9-4b28-b29a-20f9e06039a3-1409072175101@3capp-gm
x-bs56>, acbern@gmx.de writes:
Tellurium/copper is used because it machines much better than pure
copper, (20% -> 90%) without ruining the conductivity too much.
Most other metals which can improve its machinability has big
negative impacts on the conductivity of copper.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
--------
In message <trinity-9076c25b-a0a9-4b28-b29a-20f9e06039a3-1409072175101@3capp-gm
x-bs56>, acbern@gmx.de writes:
Tellurium/copper is used because it machines much better than pure
copper, (20% -> 90%) without ruining the conductivity too much.
Most other metals which can improve its machinability has big
negative impacts on the conductivity of copper.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
RE
Randy Evans
Tue, Aug 26, 2014 9:16 PM
Bill,
I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.
Randy
On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3637@att.net wrote:
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds
per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front panel
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand what
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and
get a MATH ERR symbol on the display. I tried it a couple of times and
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy at
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this sequence
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads about
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too large.
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is to
Cal Lab to charge the extra battery before they ship the 732A back
When you remove and work on the battery pack always have the
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost to
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3 mA
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4 x
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened up
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Bill,
I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.
Randy
On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold <wpgold3637@att.net> wrote:
> Not that I know of. Just wait and when you don't see a "SMPL" on the
> display it is done. But then with 1000 PLC that is around 16.66 seconds
> per
> reading times 100 readings is somewhere around 28 minutes and there is
> probably some overhead time so around 30 minutes. Not from the front panel
> at any rate.
>
> Bill
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Monday, August 25, 2014 9:23 PM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Is there any way to tell when the function key routine is complete? In
> the
> > case of taking multiple readings using the DEFKEY and MATH function, I
> > don't see any indication when the routine is complete. In one particular
> > case, I am taking a 100 readings with NLPC set for 1000 so its a long
> while
> > before it's complete, but i have to guess when it's done.
> >
> > Thanks,
> >
> > Randy
> >
> >
> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > Bill,
> > >
> > > I am trying to figure out the MATH function without much success. I
> input
> > > the sequence you said (I looked up the instructions to understand what
> you
> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
> and
> > > it shows up on the display when I input BLUE F1. I hit ENTER and it
> takes
> > > the 40 measurements and the MATH symbol shows on the display during the
> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and
> I
> > > get a MATH ERR symbol on the display. I tried it a couple of times and
> the
> > > same result so I am doing something wrong. Is there a better source
> for
> > > explaining how to do front panel masurements than the User Guide, which
> > > seems oriented at programming automatic rather than manual
> measurements.
> > >
> > > Randy
> > >
> > >
> > >
> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> > >
> > >> Randy:
> > >>
> > >> The MATH function is accessible from the keypad. I don't have an
> IEEE
> > >> interface right now that works. You can also program the numeric
> keypad
> > >> keys to have preprogrammed functions. DEFKEY
> > >>
> > >> I have made my own "low thermal" measurement leads from Pomona
> #4892
> > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at
> the
> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
> plans
> > >> to do custom cables with 16 ga. bare copper wire that I will twist and
> > >> then
> > >> put a braided shield over it. I simply cannot find what I want so I
> will
> > >> build my own cable. I have done something like this before and it
> worked
> > >> fine. When I get a "round toit".
> > >>
> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
> have
> > >> used
> > >> in the past to verify my homemade "low thermal" leads as described
> above.
> > >> Frankly I cannot see any difference between using the 1756 cables and
> my
> > >> homemade cables once I give them a few minutes for the thermals to go
> > >> away.
> > >> As far as I can tell and measure the differences, if any, are below
> 0.1
> > >> ppm
> > >> at 10 volts.
> > >>
> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
> all
> > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
> far
> > >> as
> > >> the instability of the readings it is hard to determine which is
> causing
> > >> the
> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
> is
> > >> set
> > >> the MATH to "Statistics" (store high reading/low reading/ and mean of
> the
> > >> readings) in the registers, the number of readings to "40", the
> trigger
> to
> > >> "hold" (which keeps the meter from triggering until I press "ENTER"
> and
> > >> then
> > >> trigger the sequence of 40 readings when I push the "ENTER" button.
> You
> > >> can
> > >> do all of this manually from the keypads but since I use this sequence
> a
> > >> lot
> > >> I have preprogrammed it. This is after I set digits to "8" and PLC to
> > >> 100.
> > >> Once those 40 readings are finished then you can access the various
> MATH
> > >> statistic registers, using the menu, by entering "MATH" and then a 2
> for
> > >> low, a 4 for mean, and 13 for high. Of course you could do all of
> this
> > >> through the IEEE also. The 3458A has a very rich set of measurement
> > >> commands. I am still learning all of them. It depends upon what I am
> > >> trying to accomplish.
> > >>
> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> > >> resistive dividers from the 10 volt, I don't see how they could
> contribute
> > >> to the varying readings you are measuring. I think I would put a
> short
> on
> > >> the input of the 3458A and manually set the range to 1 volt and then
> > >> observe
> > >> the variations that way without the 732A involved. When I do this I
> see a
> > >> variation from low reading to high reading of 0.125 uVolts and then
> > >> another
> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
> > >> side
> > >> of the measurment terminals, GUARD connected doesn't seem to affect
> the
> > >> readings. So that is the base noise of the 3458A without the 732A,
> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output
> I
> > >> got
> > >> a variation of 0.159 uVolts using the same 40 reading method above. I
> > >> would
> > >> use this to determine where your problem might exist. Just having the
> > >> meter
> > >> input shorted will point you in the right direction. Meter, cables or
> > >> 732A.
> > >>
> > >> Sorry for the long dissertation. Friends get mad at me for being
> so
> > >> detailed sometimes.
> > >>
> > >> Bill
> > >>
> > >> ----- Original Message -----
> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > >> Sent: Sunday, August 24, 2014 7:22 AM
> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> > >>
> > >>
> > >> > Bill,
> > >> >
> > >> > I have convinced myself that the problem I an seeing is due to
> thermals.
> > >> > If I move the cables (with gold-plated banana plugs) using a small
> > >> towel
> > >> > rather than letting my hand touch the plugs, it is much more stable.
> > >> If I
> > >> > then hold the banana plug with my hand after the reading has
> stabilized,
> > >> > the reading drifts rapidly upward. I am trying to check the
> stability
> > >> of
> > >> > the reading but I haven't figured out the MATH function yet. I
> assume
> > >> this
> > >> > is a programmed function using GPIB only?
> > >> >
> > >> > The stability I am seeing by manually recording readings (using NLPC
> of
> > >> 100
> > >> > and 1000) is much greater than what you are measuring on your
> system.
> > >> Not
> > >> > sure how to ascertain if it's the 3458A or the 732. The value of
> the
> > >> > readings are very different between the two - the 3458 reads about
> 50
> uV
> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> Rather
> > >> large
> > >> > differences (this is after an ACAL). I need to find some better
> cables
> > >> to
> > >> > make sure the errors are not due to thermals again.
> > >> >
> > >> > Randy
> > >> >
> > >> >
> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
> wrote:
> > >> >
> > >> > > Randy:
> > >> > >
> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
> fit
> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
> and
> > >> > > others
> > >> > > that are in this size and package. Order from one of the usual
> > >> electronics
> > >> > > distributors like Allied, Mouser, Digikey. This is a very common
> > >> battery
> > >> > > as
> > >> > > it is used in a lot of "EXIT" signs so they are lighted when the
> power
> > >> goes
> > >> > > out. I don't see how the 12V 7AH will fit as they are too large.
> I
> > >> guess
> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
> volt
> > >> 4
> > >> AH
> > >> > > but the terminals are in the wrong place so you will have to
> "nibble"
> > >> out
> > >> > > the aluminum plate that holds them in the 732A battery pack. You
> have
> > >> to
> > >> > > be
> > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> > >> connection
> > >> > > leads to deal with and connect correctly. I would stick with the
> 6V
> > >> 4AH.
> > >> > > New batteries will last around 12 to 14 hours before the "CAL"
> light
> > >> goes
> > >> > > out when AC power is not applied. So shipping to Cal Lab can be a
> > >> problem
> > >> > > if it is a distance away, or you have to use a shipper like UPS or
> > >> FEDEX
> > >> > > and
> > >> > > you ship the night before and then use their "Morning delivery"
> and
> > >> the
> > >> Cal
> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> course
> > >> you
> > >> > > could always strap another battery on the 732A and hook it up to
> the
> > >> "ext
> > >> > > power" plug to last longer. I have seen it done. The issue is to
> get
> > >> the
> > >> > > Cal Lab to charge the extra battery before they ship the 732A back
> to
> > >> you.
> > >> > >
> > >> > > When you remove and work on the battery pack always have the
> AC
> > >> power
> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> regulated
> > >> supply
> > >> > > is working.
> > >> > >
> > >> > > The "CAL" light is to indicate that power has not been lost to
> the
> > >> > > Reference Amp or other associated circuits. When the raw supply
> > >> (battery)
> > >> > > voltage drops below about 21 volts the "CAL" light will go out.
> Below
> > >> that
> > >> > > voltage the heater circuits will not work correctly and the 18.6
> volt
> > >> > > regulated supply will not regulate. The requirement is that the
> > >> Reference
> > >> > > Amp be kept "alive" at all times to maintain the output voltage
> that
> > >> was
> > >> > > measured at the time of the most recent Calibration or
> Certification.
> > >> When
> > >> > > the semiconductor junctions are unbiased and cool off when power
> is
> > >> lost,
> > >> > > and then power is restored the result will be a different 10 volts
> > >> than
> > >> > > before the power failure. My experience is that after all of the
> > >> years
> > >> > > that
> > >> > > these units have been powered up, this won't happen and when power
> is
> > >> lost
> > >> > > and then restored, even months later, the 732A will come back to
> > >> almost
> > >> > > exactly the same 10 volts as when they lost power, usually with in
> 0.2
> > >> PPM
> > >> > > after 24 hours of "warm up".
> > >> > >
> > >> > > What type of hookup leads are you using when measuring the 1
> volt
> > >> > > output? What is the PLC set to? I always use 100 PLC to measure
> > >> this.
> > >> If
> > >> > > you don't have "low thermal" connection leads you can experience
> uV
> > >> changes
> > >> > > for a minute or more after plugging in the leads due to the
> "thermals"
> > >> > > generated because of the difference in temperature between the
> banana
> > >> jacks
> > >> > > on the 732A and the banana plugs of the leads. I have found that
> even
> > >> just
> > >> > > plugging in the lead will generate a thermal difference because of
> > >> > > difference of temps and some heating due to the physical act of
> just
> > >> > > inserting the plug because of friction between the jack and plug
> (my
> > >> theory
> > >> > > at any rate). You have to allow at least a minute or more before
> > >> being
> > >> > > able
> > >> > > to make a measurement after plugging in the leads. I just
> measured
> > >> the
> > >> > > variation of the 1 volt output of my 732A and using my 3458A and I
> > >> got a
> > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC
> on
> > >> the
> > >> 1
> > >> > > volt range of the 3458A. Using the MATH function and all of the
> data
> > >> you
> > >> > > can collect. That was after waiting for several minutes after
> > >> plugging
> > >> in
> > >> > > the leads.
> > >> > >
> > >> > > I hope all of this helps.
> > >> > >
> > >> > > Bill
> > >> > >
> > >> > >
> > >> > >
> > >> > > ----- Original Message -----
> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > > To: "Discussion of precise voltage measurement"
> <volt-nuts@febo.com>
> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >
> > >> > >
> > >> > > > Todd,
> > >> > > >
> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> batteries
> > >> that
> > >> I
> > >> > > > keep topped off and they have very low current draw (~2 to 3 mA
> at
> > >> 13.5
> > >> > > > VDC) when charged and at their float voltage, so I am pretty
> sure
> > >> they
> > >> > > are
> > >> > > > in good condition. I will look at getting those in the units
> after
> > >> I
> > >> > > > ascertain the condition of the 732.
> > >> > > >
> > >> > > > So now I have a what appears to be a functioning 3458A and a
> 732A
> > >> but
> > >> > > they
> > >> > > > slightly disagree. I am like the man with two watches that
> disagree
> > >> on
> > >> > > the
> > >> > > > time - which is correct? For the moment, i am only concerned
> with
> > >> > > > stability. The need for absolute accuracy will come later.
> > >> > > >
> > >> > > > Randy
> > >> > > >
> > >> > > >
> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
> <tmicallef@gmail.com
> > >> >
> > >> > > wrote:
> > >> > > >
> > >> > > > > Randy,
> > >> > > > >
> > >> > > > > You have two possible choices. It can be configured with 4 x
> 6v
> > >> 4Ah
> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> owner
> > >> has
> > >> > > > > modified the battery pack already. A couple of mine needed a
> > >> nibbler
> > >> > > tool
> > >> > > > > to remove enough of the aluminum cover that fits over the tops
> of
> > >> the
> > >> > > > > batteries. The original cover will short out to the battery
> tabs
> > >> > > regardless
> > >> > > > > of the battery configuration if this is not done.
> > >> > > > >
> > >> > > > > You can find larger capacity batteries that will give you
> slightly
> > >> more
> > >> > > > > battery life. I lost a couple sets of mail-order batteries
> after a
> > >> few
> > >> > > > > extended outages. I would recommend going with locally bought
> > >> batteries
> > >> > > > > instead of the cheaper mail order. My local Batteries Plus
> will
> > >> > > typically
> > >> > > > > have some warranty if I remember correctly. Moving forward I
> will
> > >> only
> > >> > > use
> > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> > >> equalize
> > >> > > them
> > >> > > > > before putting them in the 732A. I think the cheap batteries
> did
> > >> not
> > >> > > > > discharge equally, and would not recover when power was
> applied.
> > >> > > > >
> > >> > > > > Inspect the back plane for damaged traces and look at the
> > >> capacitors. I
> > >> > > had
> > >> > > > > a few that looked questionable. So far, I have replaced all
> the
> > >> big
> > >> > > caps
> > >> > > on
> > >> > > > > the pre-regulator and regulator boards. My feeling is that
> once
> > >> these
> > >> > > go
> > >> > > > > online, they should run as long as possible between repairs.
> > >> > > > >
> > >> > > > > The battery charger circuit may need adjusting. I tweaked mine
> and
> > >> it
> > >> > > > > seemed to work fine.
> > >> > > > >
> > >> > > > > Todd
> > >> > > > >
> > >> > > > >
> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > >> > > randyevans2688@gmail.com>
> > >> > > > > wrote:
> > >> > > > >
> > >> > > > > > I received my Fluke 732A today. Just powered it up but it
> needs
> > >> new
> > >> > > > > > batteries. Any suggestions for sources (I haven't opened up
> the
> > >> unit
> > >> > > > > yet -
> > >> > > > > > I want to make sure it works before doing that). Also
> received
> > >> the
> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
> ran
> > >> data
> > >> > > > > dumper
> > >> > > > > > program to get the CAL data from my 3458A. Should be a busy
> > >> weekend.
> > >> > > > > >
> > >> > > > > > Randy
> > >> > > > > > _______________________________________________
> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > > > To unsubscribe, go to
> > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > > > and follow the instructions there.
> > >> > > > > >
> > >> > > > > _______________________________________________
> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > > To unsubscribe, go to
> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > > and follow the instructions there.
> > >> > > > >
> > >> > > > _______________________________________________
> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > To unsubscribe, go to
> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > and follow the instructions there.
> > >> > >
> > >> > > _______________________________________________
> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > To unsubscribe, go to
> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > and follow the instructions there.
> > >> > >
> > >> > _______________________________________________
> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > and follow the instructions there.
> > >>
> > >> _______________________________________________
> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> and follow the instructions there.
> > >>
> > >
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
DM
Dave M
Wed, Aug 27, 2014 12:01 AM
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC
Voltage Measurements by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Great!!@! Many Thanks!!
Dave M
acbern@gmx.de wrote:
> Dave,
>
> the title is:
> Thermal Voltage Converters and Comparator for Very Accurate AC
> Voltage Measurements by E.S.Williams.
>
> Adrian
>
>
>> Gesendet: Montag, 25. August 2014 um 19:32 Uhr
>> Von: "Dave M" <dgminala@mediacombb.net>
>> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Betreff: Re: [volt-nuts] AC calibration
>>
>> Adrian,
>> Do you have a link or title for the NIST paper that you mentioned?
>
A
acbern@gmx.de
Wed, Aug 27, 2014 12:41 AM
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only there many subsequent measuremnts with statistics make sense) I am already getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal) unless the amb. temperature is really very stable (half a degree already adds about 0.25ppm at 10v)
thanks
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In the
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long while
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and the
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
as
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only there many subsequent measuremnts with statistics make sense) I am already getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal) unless the amb. temperature is really very stable (half a degree already adds about 0.25ppm at 10v)
thanks
> Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
> Von: "Randy Evans" <randyevans2688@gmail.com>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> Is there any way to tell when the function key routine is complete? In the
> case of taking multiple readings using the DEFKEY and MATH function, I
> don't see any indication when the routine is complete. In one particular
> case, I am taking a 100 readings with NLPC set for 1000 so its a long while
> before it's complete, but i have to guess when it's done.
>
> Thanks,
>
> Randy
>
>
> On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > Bill,
> >
> > I am trying to figure out the MATH function without much success. I input
> > the sequence you said (I looked up the instructions to understand what you
> > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
> > it shows up on the display when I input BLUE F1. I hit ENTER and it takes
> > the 40 measurements and the MATH symbol shows on the display during the
> > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> > get a MATH ERR symbol on the display. I tried it a couple of times and the
> > same result so I am doing something wrong. Is there a better source for
> > explaining how to do front panel masurements than the User Guide, which
> > seems oriented at programming automatic rather than manual measurements.
> >
> > Randy
> >
> >
> >
> > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> >
> >> Randy:
> >>
> >> The MATH function is accessible from the keypad. I don't have an IEEE
> >> interface right now that works. You can also program the numeric keypad
> >> keys to have preprogrammed functions. DEFKEY
> >>
> >> I have made my own "low thermal" measurement leads from Pomona #4892
> >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
> >> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
> >> to do custom cables with 16 ga. bare copper wire that I will twist and
> >> then
> >> put a braided shield over it. I simply cannot find what I want so I will
> >> build my own cable. I have done something like this before and it worked
> >> fine. When I get a "round toit".
> >>
> >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> >> used
> >> in the past to verify my homemade "low thermal" leads as described above.
> >> Frankly I cannot see any difference between using the 1756 cables and my
> >> homemade cables once I give them a few minutes for the thermals to go
> >> away.
> >> As far as I can tell and measure the differences, if any, are below 0.1
> >> ppm
> >> at 10 volts.
> >>
> >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
> >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
> >> as
> >> the instability of the readings it is hard to determine which is causing
> >> the
> >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
> >> set
> >> the MATH to "Statistics" (store high reading/low reading/ and mean of the
> >> readings) in the registers, the number of readings to "40", the trigger to
> >> "hold" (which keeps the meter from triggering until I press "ENTER" and
> >> then
> >> trigger the sequence of 40 readings when I push the "ENTER" button. You
> >> can
> >> do all of this manually from the keypads but since I use this sequence a
> >> lot
> >> I have preprogrammed it. This is after I set digits to "8" and PLC to
> >> 100.
> >> Once those 40 readings are finished then you can access the various MATH
> >> statistic registers, using the menu, by entering "MATH" and then a 2 for
> >> low, a 4 for mean, and 13 for high. Of course you could do all of this
> >> through the IEEE also. The 3458A has a very rich set of measurement
> >> commands. I am still learning all of them. It depends upon what I am
> >> trying to accomplish.
> >>
> >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> >> resistive dividers from the 10 volt, I don't see how they could contribute
> >> to the varying readings you are measuring. I think I would put a short on
> >> the input of the 3458A and manually set the range to 1 volt and then
> >> observe
> >> the variations that way without the 732A involved. When I do this I see a
> >> variation from low reading to high reading of 0.125 uVolts and then
> >> another
> >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
> >> side
> >> of the measurment terminals, GUARD connected doesn't seem to affect the
> >> readings. So that is the base noise of the 3458A without the 732A,
> >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
> >> got
> >> a variation of 0.159 uVolts using the same 40 reading method above. I
> >> would
> >> use this to determine where your problem might exist. Just having the
> >> meter
> >> input shorted will point you in the right direction. Meter, cables or
> >> 732A.
> >>
> >> Sorry for the long dissertation. Friends get mad at me for being so
> >> detailed sometimes.
> >>
> >> Bill
> >>
> >> ----- Original Message -----
> >> From: "Randy Evans" <randyevans2688@gmail.com>
> >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> Sent: Sunday, August 24, 2014 7:22 AM
> >> Subject: Re: [volt-nuts] 732A and Prologix received
> >>
> >>
> >> > Bill,
> >> >
> >> > I have convinced myself that the problem I an seeing is due to thermals.
> >> > If I move the cables (with gold-plated banana plugs) using a small
> >> towel
> >> > rather than letting my hand touch the plugs, it is much more stable.
> >> If I
> >> > then hold the banana plug with my hand after the reading has stabilized,
> >> > the reading drifts rapidly upward. I am trying to check the stability
> >> of
> >> > the reading but I haven't figured out the MATH function yet. I assume
> >> this
> >> > is a programmed function using GPIB only?
> >> >
> >> > The stability I am seeing by manually recording readings (using NLPC of
> >> 100
> >> > and 1000) is much greater than what you are measuring on your system.
> >> Not
> >> > sure how to ascertain if it's the 3458A or the 732. The value of the
> >> > readings are very different between the two - the 3458 reads about 50 uV
> >> > high on the 10 V output and about 12 uV low on the 1V output. Rather
> >> large
> >> > differences (this is after an ACAL). I need to find some better cables
> >> to
> >> > make sure the errors are not due to thermals again.
> >> >
> >> > Randy
> >> >
> >> >
> >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
> >> >
> >> > > Randy:
> >> > >
> >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> >> > > others
> >> > > that are in this size and package. Order from one of the usual
> >> electronics
> >> > > distributors like Allied, Mouser, Digikey. This is a very common
> >> battery
> >> > > as
> >> > > it is used in a lot of "EXIT" signs so they are lighted when the power
> >> goes
> >> > > out. I don't see how the 12V 7AH will fit as they are too large. I
> >> guess
> >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt
> >> 4
> >> AH
> >> > > but the terminals are in the wrong place so you will have to "nibble"
> >> out
> >> > > the aluminum plate that holds them in the 732A battery pack. You have
> >> to
> >> > > be
> >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> >> connection
> >> > > leads to deal with and connect correctly. I would stick with the 6V
> >> 4AH.
> >> > > New batteries will last around 12 to 14 hours before the "CAL" light
> >> goes
> >> > > out when AC power is not applied. So shipping to Cal Lab can be a
> >> problem
> >> > > if it is a distance away, or you have to use a shipper like UPS or
> >> FEDEX
> >> > > and
> >> > > you ship the night before and then use their "Morning delivery" and
> >> the
> >> Cal
> >> > > Lab is expecting your 732A. Same on the way back to you. Of course
> >> you
> >> > > could always strap another battery on the 732A and hook it up to the
> >> "ext
> >> > > power" plug to last longer. I have seen it done. The issue is to get
> >> the
> >> > > Cal Lab to charge the extra battery before they ship the 732A back to
> >> you.
> >> > >
> >> > > When you remove and work on the battery pack always have the AC
> >> power
> >> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> >> supply
> >> > > is working.
> >> > >
> >> > > The "CAL" light is to indicate that power has not been lost to the
> >> > > Reference Amp or other associated circuits. When the raw supply
> >> (battery)
> >> > > voltage drops below about 21 volts the "CAL" light will go out. Below
> >> that
> >> > > voltage the heater circuits will not work correctly and the 18.6 volt
> >> > > regulated supply will not regulate. The requirement is that the
> >> Reference
> >> > > Amp be kept "alive" at all times to maintain the output voltage that
> >> was
> >> > > measured at the time of the most recent Calibration or Certification.
> >> When
> >> > > the semiconductor junctions are unbiased and cool off when power is
> >> lost,
> >> > > and then power is restored the result will be a different 10 volts
> >> than
> >> > > before the power failure. My experience is that after all of the
> >> years
> >> > > that
> >> > > these units have been powered up, this won't happen and when power is
> >> lost
> >> > > and then restored, even months later, the 732A will come back to
> >> almost
> >> > > exactly the same 10 volts as when they lost power, usually with in 0.2
> >> PPM
> >> > > after 24 hours of "warm up".
> >> > >
> >> > > What type of hookup leads are you using when measuring the 1 volt
> >> > > output? What is the PLC set to? I always use 100 PLC to measure
> >> this.
> >> If
> >> > > you don't have "low thermal" connection leads you can experience uV
> >> changes
> >> > > for a minute or more after plugging in the leads due to the "thermals"
> >> > > generated because of the difference in temperature between the banana
> >> jacks
> >> > > on the 732A and the banana plugs of the leads. I have found that even
> >> just
> >> > > plugging in the lead will generate a thermal difference because of
> >> > > difference of temps and some heating due to the physical act of just
> >> > > inserting the plug because of friction between the jack and plug (my
> >> theory
> >> > > at any rate). You have to allow at least a minute or more before
> >> being
> >> > > able
> >> > > to make a measurement after plugging in the leads. I just measured
> >> the
> >> > > variation of the 1 volt output of my 732A and using my 3458A and I
> >> got a
> >> > > total difference of 0.159 uV over 40 measurements using 100 PLC on
> >> the
> >> 1
> >> > > volt range of the 3458A. Using the MATH function and all of the data
> >> you
> >> > > can collect. That was after waiting for several minutes after
> >> plugging
> >> in
> >> > > the leads.
> >> > >
> >> > > I hope all of this helps.
> >> > >
> >> > > Bill
> >> > >
> >> > >
> >> > >
> >> > > ----- Original Message -----
> >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> >> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> > > Sent: Friday, August 22, 2014 9:03 PM
> >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >
> >> > >
> >> > > > Todd,
> >> > > >
> >> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> >> that
> >> I
> >> > > > keep topped off and they have very low current draw (~2 to 3 mA at
> >> 13.5
> >> > > > VDC) when charged and at their float voltage, so I am pretty sure
> >> they
> >> > > are
> >> > > > in good condition. I will look at getting those in the units after
> >> I
> >> > > > ascertain the condition of the 732.
> >> > > >
> >> > > > So now I have a what appears to be a functioning 3458A and a 732A
> >> but
> >> > > they
> >> > > > slightly disagree. I am like the man with two watches that disagree
> >> on
> >> > > the
> >> > > > time - which is correct? For the moment, i am only concerned with
> >> > > > stability. The need for absolute accuracy will come later.
> >> > > >
> >> > > > Randy
> >> > > >
> >> > > >
> >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
> >> >
> >> > > wrote:
> >> > > >
> >> > > > > Randy,
> >> > > > >
> >> > > > > You have two possible choices. It can be configured with 4 x 6v
> >> 4Ah
> >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> >> has
> >> > > > > modified the battery pack already. A couple of mine needed a
> >> nibbler
> >> > > tool
> >> > > > > to remove enough of the aluminum cover that fits over the tops of
> >> the
> >> > > > > batteries. The original cover will short out to the battery tabs
> >> > > regardless
> >> > > > > of the battery configuration if this is not done.
> >> > > > >
> >> > > > > You can find larger capacity batteries that will give you slightly
> >> more
> >> > > > > battery life. I lost a couple sets of mail-order batteries after a
> >> few
> >> > > > > extended outages. I would recommend going with locally bought
> >> batteries
> >> > > > > instead of the cheaper mail order. My local Batteries Plus will
> >> > > typically
> >> > > > > have some warranty if I remember correctly. Moving forward I will
> >> only
> >> > > use
> >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> >> equalize
> >> > > them
> >> > > > > before putting them in the 732A. I think the cheap batteries did
> >> not
> >> > > > > discharge equally, and would not recover when power was applied.
> >> > > > >
> >> > > > > Inspect the back plane for damaged traces and look at the
> >> capacitors. I
> >> > > had
> >> > > > > a few that looked questionable. So far, I have replaced all the
> >> big
> >> > > caps
> >> > > on
> >> > > > > the pre-regulator and regulator boards. My feeling is that once
> >> these
> >> > > go
> >> > > > > online, they should run as long as possible between repairs.
> >> > > > >
> >> > > > > The battery charger circuit may need adjusting. I tweaked mine and
> >> it
> >> > > > > seemed to work fine.
> >> > > > >
> >> > > > > Todd
> >> > > > >
> >> > > > >
> >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> >> > > randyevans2688@gmail.com>
> >> > > > > wrote:
> >> > > > >
> >> > > > > > I received my Fluke 732A today. Just powered it up but it needs
> >> new
> >> > > > > > batteries. Any suggestions for sources (I haven't opened up the
> >> unit
> >> > > > > yet -
> >> > > > > > I want to make sure it works before doing that). Also received
> >> the
> >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> >> data
> >> > > > > dumper
> >> > > > > > program to get the CAL data from my 3458A. Should be a busy
> >> weekend.
> >> > > > > >
> >> > > > > > Randy
> >> > > > > > _______________________________________________
> >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > > > To unsubscribe, go to
> >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > > > and follow the instructions there.
> >> > > > > >
> >> > > > > _______________________________________________
> >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > > To unsubscribe, go to
> >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > > and follow the instructions there.
> >> > > > >
> >> > > > _______________________________________________
> >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > To unsubscribe, go to
> >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > and follow the instructions there.
> >> > >
> >> > > _______________________________________________
> >> > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > To unsubscribe, go to
> >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > and follow the instructions there.
> >> > >
> >> > _______________________________________________
> >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > and follow the instructions there.
> >>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >>
> >
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
P
pa4tim@gmail.com
Wed, Aug 27, 2014 5:08 AM
Thanks, I did not new NIST has so much interesting information on their site.
I read the following on their site :
"multirange instruments with up to eight decimal digits of adjustability, are not considered by NIST to be standards "
Does that mean a HP3458 ? Fluke in the Netherlands used a HP-3458 and a 732 to calibrate the Fluke 5101 that I'm working on at this moment (it failed calibration due to some hardware faults)
Everybody thanks for the information. Turns out, a friend has a GR 1455AH for me, that is some kind a AC KV divider.
I will test if the output of my HP 3400 is usable to connect a DMM. I have one I restored a few years ago. Besides that someone mentioned to look at the LT-1088.
But that has to wait until I have some more time. I only repaired calibration, percission gear and RF stuff for my hobby, but, not planned, I started to do this on a commercial base too. See my (non commercial) site about my collection calibration and other gear and projects www.pa4tim.nl .
Today the 5101 goes back to the custommer but he brings 2 other instruments in need of some TLC and a precheck before they are shipped to Fluke for calibration. And yesterday an other company asked if I want/can repair a G&M safety tester (a sort of megger on steroids) that died during calibration.
Fred
Verzonden met Windows Mail
Van: acbern@gmx.de
Verzonden: ‎dinsdag‎ ‎26‎ ‎augustus‎ ‎2014 ‎18‎:‎56
Aan: volt-nuts
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Dave M
acbern@gmx.de wrote:
fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be <10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.
adrian
Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer. Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer. At that point,
the AC voltage source is equal to that of the DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you
have to have a converter for each range of voltages that you need to
measure. The thermal converters used with this type of transfer
standard isn't great (50 MHz or so typical), but their accuracy far
surpasses that of the thermistor type sensors.
There are other brands and models of thermal transfer standards, but
I have a Fluke model 540 and a few thermal converters. That's why I
referred you to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two
AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for
1V, 10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
Thanks, I did not new NIST has so much interesting information on their site.
I read the following on their site :
"multirange instruments with up to eight decimal digits of adjustability, are not considered by NIST to be standards "
Does that mean a HP3458 ? Fluke in the Netherlands used a HP-3458 and a 732 to calibrate the Fluke 5101 that I'm working on at this moment (it failed calibration due to some hardware faults)
Everybody thanks for the information. Turns out, a friend has a GR 1455AH for me, that is some kind a AC KV divider.
I will test if the output of my HP 3400 is usable to connect a DMM. I have one I restored a few years ago. Besides that someone mentioned to look at the LT-1088.
But that has to wait until I have some more time. I only repaired calibration, percission gear and RF stuff for my hobby, but, not planned, I started to do this on a commercial base too. See my (non commercial) site about my collection calibration and other gear and projects www.pa4tim.nl .
Today the 5101 goes back to the custommer but he brings 2 other instruments in need of some TLC and a precheck before they are shipped to Fluke for calibration. And yesterday an other company asked if I want/can repair a G&M safety tester (a sort of megger on steroids) that died during calibration.
Fred
Verzonden met Windows Mail
Van: acbern@gmx.de
Verzonden: ‎dinsdag‎ ‎26‎ ‎augustus‎ ‎2014 ‎18‎:‎56
Aan: volt-nuts
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.
Adrian
> Gesendet: Montag, 25. August 2014 um 19:32 Uhr
> Von: "Dave M" <dgminala@mediacombb.net>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] AC calibration
>
> Adrian,
> Do you have a link or title for the NIST paper that you mentioned?
>
> Dave M
>
>
> acbern@gmx.de wrote:
> > fred,
> > generally you raise a good point, I had the same issue of calibrating
> > an ac voltage to a high level of accuracy. you need this e.g. to
> > validate the self.cal of a 3458a or other precison stuff like the
> > 8506a0.
> >
> > what i would recommend to do if you want to keep costs down is:
> > in a nutshell, get a thermal converter in the lowest range you need
> > and a second one on range above. build a set of resistor range
> > extenders (rf type with appropriate connectors and housings) to
> > expand the range to where you need to be max. get one of the thermal
> > converter calibrated (the higher one usually, and you need to havr
> > good cal lab, should be <10ppm accuracy) and use it to calibrate the
> > rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
> > thermal converters! at higher frequencies, due to reflections and
> > stray capacitance/inductance influences, the accuracy decreases. the
> > resistor range extenders though, if build up correctly, only have a
> > few ppm impact (there is a paper from nist on that, but this is only
> > typical). you can calibrate all converters to the one you got
> > externally calibrated. do some research in the web, when you do the
> > calibration, you need to determine the so-called constant N. then do
> > an ac, dc+, ac, dc-, ac measurement between the the two and establish
> > the deviation, also establish the error propagation. the end result
> > will be a set of highly precise (low inaccuracies9 thermal converters
> > good enough to calibrate a 3458a an better devices. if you want to
> > spend the money, you could also buy a set of converters/range
> > resistors (with/without a 540), that typically is a few k altogether,
> > while a single device sometimes is available for below 100 bucks. you
> > need to have a stable 7.5 digit nanovoltmeter though for the
> > measurements of the tvcs (34420a or 2182 typically ) and precision
> > (stable) dc and ac sources. but in the end, all you need is a single
> > calibrated thermal converter.
> >
> > adrian
> >
> >
> >
> >> Gesendet: Montag, 25. August 2014 um 18:38 Uhr
> >> Von: "Dave M" <dgminala@mediacombb.net>
> >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> Betreff: Re: [volt-nuts] AC calibration
> >>
> >> Well, you sort of answered your own question. The equipment is
> >> called a Thermal Transfer Standard, but instead of thermistors, it
> >> uses a thermocouple. Look at the manual for the Fluke 540B
> >> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
> >> it's done. Basically, the AC source is input into the transfer
> >> standard, and the standard's internal reference voltage is adjusted
> >> for a null on the galvanometer. Leaving the reference voltage
> >> setting alone, a DC voltage is input into the unit, and the DC
> >> source is adjusted for a null on the galvanometer. At that point,
> >> the AC voltage source is equal to that of the DC voltage source.
> >>
> >> Ther are thermocouple-type thermal converters used for RF voltage
> >> measurements with the transfer standard. They aren't cheap, and you
> >> have to have a converter for each range of voltages that you need to
> >> measure. The thermal converters used with this type of transfer
> >> standard isn't great (50 MHz or so typical), but their accuracy far
> >> surpasses that of the thermistor type sensors.
> >>
> >> There are other brands and models of thermal transfer standards, but
> >> I have a Fluke model 540 and a few thermal converters. That's why I
> >> referred you to the manual for it.
> >>
> >> Cheers,
> >> Dave M
> >>
> >>
> >> pa4tim@gmail.com wrote:
> >>> Is there a way to link an AC voltage to a DC source for compare. I
> >>> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
> >>> against standardcells. But for AC I can not do that. I have two
> >>> AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
> >>> ago.
> >>>
> >>> My idea is in theory simple. It is based on the thermal converters
> >>> used in RF powermeters. Two resistors, two high resolution
> >>> temperature meters. AC on the first en DC on the second. If both are
> >>> the same temperature the AC voltage is the same as the DC voltage.
> >>> But I'm sure some people here have done this in the past. I would
> >>> like to use it for 50 to 100 kHz (or less) and something like for
> >>> 1V, 10V and 100V (and use several resistors/heaters.)
> >>>
> >>> Or mabey there is an other way to convert AC (for RF it can be done
> >>> with lightbubs but I never tryed that) I do not mind if it is slow
> >>> etc, I like this sort of experiments. You can learn a lot from it.
> >>>
> >>> Fred, pa4tim
> >>
> >>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
> >> the instructions there.
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> When the people fear the government, there is tyranny. When the
> government fears the people, there is liberty -- Thomas Jefferson
>
>
> Dave M
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
A
acbern@gmx.de
Wed, Aug 27, 2014 9:14 AM
what nist means is that a precision meter is not considered a standard. you always measure against a true standard (732a, esi sr104...). nist does not mean that as part of doing equipment calibration a 3458a cannot be used as aid. also keep in mind nist has a different approach than a cal lab, for obvious reasons. a cal lab would do a transfer measurement using a 3458a and a voltage/current source, sure.
there are precision ratio transformers available from various vendors, they can be used for low frequency precision calibrations, up to 20khz only (if anybodys knows one specified above 20khz I am interested to hear). so they will help you only partially in calibrating even a low precision 5101.
Gesendet: Mittwoch, 27. August 2014 um 07:08 Uhr
Von: pa4tim@gmail.com
An: volt-nuts volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Thanks, I did not new NIST has so much interesting information on their site.
I read the following on their site :
"multirange instruments with up to eight decimal digits of adjustability, are not considered by NIST to be standards "
Does that mean a HP3458 ? Fluke in the Netherlands used a HP-3458 and a 732 to calibrate the Fluke 5101 that I'm working on at this moment (it failed calibration due to some hardware faults)
Everybody thanks for the information. Turns out, a friend has a GR 1455AH for me, that is some kind a AC KV divider.
I will test if the output of my HP 3400 is usable to connect a DMM. I have one I restored a few years ago. Besides that someone mentioned to look at the LT-1088.
But that has to wait until I have some more time. I only repaired calibration, percission gear and RF stuff for my hobby, but, not planned, I started to do this on a commercial base too. See my (non commercial) site about my collection calibration and other gear and projects www.pa4tim.nl .
Today the 5101 goes back to the custommer but he brings 2 other instruments in need of some TLC and a precheck before they are shipped to Fluke for calibration. And yesterday an other company asked if I want/can repair a G&M safety tester (a sort of megger on steroids) that died during calibration.
Fred
what nist means is that a precision meter is not considered a standard. you always measure against a true standard (732a, esi sr104...). nist does not mean that as part of doing equipment calibration a 3458a cannot be used as aid. also keep in mind nist has a different approach than a cal lab, for obvious reasons. a cal lab would do a transfer measurement using a 3458a and a voltage/current source, sure.
there are precision ratio transformers available from various vendors, they can be used for low frequency precision calibrations, up to 20khz only (if anybodys knows one specified above 20khz I am interested to hear). so they will help you only partially in calibrating even a low precision 5101.
> Gesendet: Mittwoch, 27. August 2014 um 07:08 Uhr
> Von: pa4tim@gmail.com
> An: volt-nuts <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] AC calibration
>
> Thanks, I did not new NIST has so much interesting information on their site.
>
>
>
>
> I read the following on their site :
>
> "multirange instruments with up to eight decimal digits of adjustability, are not considered by NIST to be standards "
>
>
>
>
> Does that mean a HP3458 ? Fluke in the Netherlands used a HP-3458 and a 732 to calibrate the Fluke 5101 that I'm working on at this moment (it failed calibration due to some hardware faults)
>
>
>
>
> Everybody thanks for the information. Turns out, a friend has a GR 1455AH for me, that is some kind a AC KV divider.
>
> I will test if the output of my HP 3400 is usable to connect a DMM. I have one I restored a few years ago. Besides that someone mentioned to look at the LT-1088.
>
>
> But that has to wait until I have some more time. I only repaired calibration, percission gear and RF stuff for my hobby, but, not planned, I started to do this on a commercial base too. See my (non commercial) site about my collection calibration and other gear and projects www.pa4tim.nl .
>
> Today the 5101 goes back to the custommer but he brings 2 other instruments in need of some TLC and a precheck before they are shipped to Fluke for calibration. And yesterday an other company asked if I want/can repair a G&M safety tester (a sort of megger on steroids) that died during calibration.
>
>
> Fred
>
>
RE
Randy Evans
Wed, Aug 27, 2014 12:13 PM
I am doing multiple 100 measurements simply to characterize the stability
of the 3458A and 732A units I just bought. After about 10 measurement sets
over 2 days I am seeing a variance of about .5 uV for the 10V output, or
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, acbern@gmx.de wrote:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am already
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal)
unless the amb. temperature is really very stable (half a degree already
adds about 0.25ppm at 10v)
thanks
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand what
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and
get a MATH ERR symbol on the display. I tried it a couple of times
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads about
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have the
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3 mA
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
Randy,
You have two possible choices. It can be configured with 4 x
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
I am doing multiple 100 measurements simply to characterize the stability
of the 3458A and 732A units I just bought. After about 10 measurement sets
over 2 days I am seeing a variance of about .5 uV for the 10V output, or
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, <acbern@gmx.de> wrote:
> hi randy,
>
> just for curiosity, why doing 100 measurements at nplc 1000. is this to
> sample a changing value?
> when i am doing 10 measurements from a stable signal at nplc 100 (only
> there many subsequent measuremnts with statistics make sense) I am already
> getting a stanard deviation below 0.1ppm.
> in a 30 minute test cycle, i would also be concerned about drifts (acal)
> unless the amb. temperature is really very stable (half a degree already
> adds about 0.25ppm at 10v)
>
> thanks
>
>
>
>
> > Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
> > Von: "Randy Evans" <randyevans2688@gmail.com>
> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Betreff: Re: [volt-nuts] 732A and Prologix received
> >
> > Is there any way to tell when the function key routine is complete? In
> the
> > case of taking multiple readings using the DEFKEY and MATH function, I
> > don't see any indication when the routine is complete. In one particular
> > case, I am taking a 100 readings with NLPC set for 1000 so its a long
> while
> > before it's complete, but i have to guess when it's done.
> >
> > Thanks,
> >
> > Randy
> >
> >
> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > Bill,
> > >
> > > I am trying to figure out the MATH function without much success. I
> input
> > > the sequence you said (I looked up the instructions to understand what
> you
> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
> 4;TRIG; and
> > > it shows up on the display when I input BLUE F1. I hit ENTER and it
> takes
> > > the 40 measurements and the MATH symbol shows on the display during the
> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and
> I
> > > get a MATH ERR symbol on the display. I tried it a couple of times
> and the
> > > same result so I am doing something wrong. Is there a better source
> for
> > > explaining how to do front panel masurements than the User Guide, which
> > > seems oriented at programming automatic rather than manual
> measurements.
> > >
> > > Randy
> > >
> > >
> > >
> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> > >
> > >> Randy:
> > >>
> > >> The MATH function is accessible from the keypad. I don't have an
> IEEE
> > >> interface right now that works. You can also program the numeric
> keypad
> > >> keys to have preprogrammed functions. DEFKEY
> > >>
> > >> I have made my own "low thermal" measurement leads from Pomona
> #4892
> > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy
> at the
> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
> plans
> > >> to do custom cables with 16 ga. bare copper wire that I will twist and
> > >> then
> > >> put a braided shield over it. I simply cannot find what I want so I
> will
> > >> build my own cable. I have done something like this before and it
> worked
> > >> fine. When I get a "round toit".
> > >>
> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
> have
> > >> used
> > >> in the past to verify my homemade "low thermal" leads as described
> above.
> > >> Frankly I cannot see any difference between using the 1756 cables and
> my
> > >> homemade cables once I give them a few minutes for the thermals to go
> > >> away.
> > >> As far as I can tell and measure the differences, if any, are below
> 0.1
> > >> ppm
> > >> at 10 volts.
> > >>
> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
> are all
> > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
> far
> > >> as
> > >> the instability of the readings it is hard to determine which is
> causing
> > >> the
> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
> is
> > >> set
> > >> the MATH to "Statistics" (store high reading/low reading/ and mean of
> the
> > >> readings) in the registers, the number of readings to "40", the
> trigger to
> > >> "hold" (which keeps the meter from triggering until I press "ENTER"
> and
> > >> then
> > >> trigger the sequence of 40 readings when I push the "ENTER" button.
> You
> > >> can
> > >> do all of this manually from the keypads but since I use this
> sequence a
> > >> lot
> > >> I have preprogrammed it. This is after I set digits to "8" and PLC to
> > >> 100.
> > >> Once those 40 readings are finished then you can access the various
> MATH
> > >> statistic registers, using the menu, by entering "MATH" and then a 2
> for
> > >> low, a 4 for mean, and 13 for high. Of course you could do all of
> this
> > >> through the IEEE also. The 3458A has a very rich set of measurement
> > >> commands. I am still learning all of them. It depends upon what I am
> > >> trying to accomplish.
> > >>
> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> > >> resistive dividers from the 10 volt, I don't see how they could
> contribute
> > >> to the varying readings you are measuring. I think I would put a
> short on
> > >> the input of the 3458A and manually set the range to 1 volt and then
> > >> observe
> > >> the variations that way without the 732A involved. When I do this I
> see a
> > >> variation from low reading to high reading of 0.125 uVolts and then
> > >> another
> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
> > >> side
> > >> of the measurment terminals, GUARD connected doesn't seem to affect
> the
> > >> readings. So that is the base noise of the 3458A without the 732A,
> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output
> I
> > >> got
> > >> a variation of 0.159 uVolts using the same 40 reading method above. I
> > >> would
> > >> use this to determine where your problem might exist. Just having the
> > >> meter
> > >> input shorted will point you in the right direction. Meter, cables or
> > >> 732A.
> > >>
> > >> Sorry for the long dissertation. Friends get mad at me for being
> so
> > >> detailed sometimes.
> > >>
> > >> Bill
> > >>
> > >> ----- Original Message -----
> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > >> Sent: Sunday, August 24, 2014 7:22 AM
> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> > >>
> > >>
> > >> > Bill,
> > >> >
> > >> > I have convinced myself that the problem I an seeing is due to
> thermals.
> > >> > If I move the cables (with gold-plated banana plugs) using a small
> > >> towel
> > >> > rather than letting my hand touch the plugs, it is much more stable.
> > >> If I
> > >> > then hold the banana plug with my hand after the reading has
> stabilized,
> > >> > the reading drifts rapidly upward. I am trying to check the
> stability
> > >> of
> > >> > the reading but I haven't figured out the MATH function yet. I
> assume
> > >> this
> > >> > is a programmed function using GPIB only?
> > >> >
> > >> > The stability I am seeing by manually recording readings (using
> NLPC of
> > >> 100
> > >> > and 1000) is much greater than what you are measuring on your
> system.
> > >> Not
> > >> > sure how to ascertain if it's the 3458A or the 732. The value of
> the
> > >> > readings are very different between the two - the 3458 reads about
> 50 uV
> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> Rather
> > >> large
> > >> > differences (this is after an ACAL). I need to find some better
> cables
> > >> to
> > >> > make sure the errors are not due to thermals again.
> > >> >
> > >> > Randy
> > >> >
> > >> >
> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
> wrote:
> > >> >
> > >> > > Randy:
> > >> > >
> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
> fit
> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
> and
> > >> > > others
> > >> > > that are in this size and package. Order from one of the usual
> > >> electronics
> > >> > > distributors like Allied, Mouser, Digikey. This is a very common
> > >> battery
> > >> > > as
> > >> > > it is used in a lot of "EXIT" signs so they are lighted when the
> power
> > >> goes
> > >> > > out. I don't see how the 12V 7AH will fit as they are too
> large. I
> > >> guess
> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
> volt
> > >> 4
> > >> AH
> > >> > > but the terminals are in the wrong place so you will have to
> "nibble"
> > >> out
> > >> > > the aluminum plate that holds them in the 732A battery pack. You
> have
> > >> to
> > >> > > be
> > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> > >> connection
> > >> > > leads to deal with and connect correctly. I would stick with the
> 6V
> > >> 4AH.
> > >> > > New batteries will last around 12 to 14 hours before the "CAL"
> light
> > >> goes
> > >> > > out when AC power is not applied. So shipping to Cal Lab can be a
> > >> problem
> > >> > > if it is a distance away, or you have to use a shipper like UPS or
> > >> FEDEX
> > >> > > and
> > >> > > you ship the night before and then use their "Morning delivery"
> and
> > >> the
> > >> Cal
> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> course
> > >> you
> > >> > > could always strap another battery on the 732A and hook it up to
> the
> > >> "ext
> > >> > > power" plug to last longer. I have seen it done. The issue is
> to get
> > >> the
> > >> > > Cal Lab to charge the extra battery before they ship the 732A
> back to
> > >> you.
> > >> > >
> > >> > > When you remove and work on the battery pack always have the
> AC
> > >> power
> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> regulated
> > >> supply
> > >> > > is working.
> > >> > >
> > >> > > The "CAL" light is to indicate that power has not been lost
> to the
> > >> > > Reference Amp or other associated circuits. When the raw supply
> > >> (battery)
> > >> > > voltage drops below about 21 volts the "CAL" light will go out.
> Below
> > >> that
> > >> > > voltage the heater circuits will not work correctly and the 18.6
> volt
> > >> > > regulated supply will not regulate. The requirement is that the
> > >> Reference
> > >> > > Amp be kept "alive" at all times to maintain the output voltage
> that
> > >> was
> > >> > > measured at the time of the most recent Calibration or
> Certification.
> > >> When
> > >> > > the semiconductor junctions are unbiased and cool off when power
> is
> > >> lost,
> > >> > > and then power is restored the result will be a different 10 volts
> > >> than
> > >> > > before the power failure. My experience is that after all of the
> > >> years
> > >> > > that
> > >> > > these units have been powered up, this won't happen and when
> power is
> > >> lost
> > >> > > and then restored, even months later, the 732A will come back to
> > >> almost
> > >> > > exactly the same 10 volts as when they lost power, usually with
> in 0.2
> > >> PPM
> > >> > > after 24 hours of "warm up".
> > >> > >
> > >> > > What type of hookup leads are you using when measuring the 1
> volt
> > >> > > output? What is the PLC set to? I always use 100 PLC to measure
> > >> this.
> > >> If
> > >> > > you don't have "low thermal" connection leads you can experience
> uV
> > >> changes
> > >> > > for a minute or more after plugging in the leads due to the
> "thermals"
> > >> > > generated because of the difference in temperature between the
> banana
> > >> jacks
> > >> > > on the 732A and the banana plugs of the leads. I have found that
> even
> > >> just
> > >> > > plugging in the lead will generate a thermal difference because of
> > >> > > difference of temps and some heating due to the physical act of
> just
> > >> > > inserting the plug because of friction between the jack and plug
> (my
> > >> theory
> > >> > > at any rate). You have to allow at least a minute or more before
> > >> being
> > >> > > able
> > >> > > to make a measurement after plugging in the leads. I just
> measured
> > >> the
> > >> > > variation of the 1 volt output of my 732A and using my 3458A and I
> > >> got a
> > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC
> on
> > >> the
> > >> 1
> > >> > > volt range of the 3458A. Using the MATH function and all of the
> data
> > >> you
> > >> > > can collect. That was after waiting for several minutes after
> > >> plugging
> > >> in
> > >> > > the leads.
> > >> > >
> > >> > > I hope all of this helps.
> > >> > >
> > >> > > Bill
> > >> > >
> > >> > >
> > >> > >
> > >> > > ----- Original Message -----
> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > > To: "Discussion of precise voltage measurement" <
> volt-nuts@febo.com>
> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >
> > >> > >
> > >> > > > Todd,
> > >> > > >
> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> batteries
> > >> that
> > >> I
> > >> > > > keep topped off and they have very low current draw (~2 to 3 mA
> at
> > >> 13.5
> > >> > > > VDC) when charged and at their float voltage, so I am pretty
> sure
> > >> they
> > >> > > are
> > >> > > > in good condition. I will look at getting those in the units
> after
> > >> I
> > >> > > > ascertain the condition of the 732.
> > >> > > >
> > >> > > > So now I have a what appears to be a functioning 3458A and a
> 732A
> > >> but
> > >> > > they
> > >> > > > slightly disagree. I am like the man with two watches that
> disagree
> > >> on
> > >> > > the
> > >> > > > time - which is correct? For the moment, i am only concerned
> with
> > >> > > > stability. The need for absolute accuracy will come later.
> > >> > > >
> > >> > > > Randy
> > >> > > >
> > >> > > >
> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
> tmicallef@gmail.com
> > >> >
> > >> > > wrote:
> > >> > > >
> > >> > > > > Randy,
> > >> > > > >
> > >> > > > > You have two possible choices. It can be configured with 4 x
> 6v
> > >> 4Ah
> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> owner
> > >> has
> > >> > > > > modified the battery pack already. A couple of mine needed a
> > >> nibbler
> > >> > > tool
> > >> > > > > to remove enough of the aluminum cover that fits over the
> tops of
> > >> the
> > >> > > > > batteries. The original cover will short out to the battery
> tabs
> > >> > > regardless
> > >> > > > > of the battery configuration if this is not done.
> > >> > > > >
> > >> > > > > You can find larger capacity batteries that will give you
> slightly
> > >> more
> > >> > > > > battery life. I lost a couple sets of mail-order batteries
> after a
> > >> few
> > >> > > > > extended outages. I would recommend going with locally bought
> > >> batteries
> > >> > > > > instead of the cheaper mail order. My local Batteries Plus
> will
> > >> > > typically
> > >> > > > > have some warranty if I remember correctly. Moving forward I
> will
> > >> only
> > >> > > use
> > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> > >> equalize
> > >> > > them
> > >> > > > > before putting them in the 732A. I think the cheap batteries
> did
> > >> not
> > >> > > > > discharge equally, and would not recover when power was
> applied.
> > >> > > > >
> > >> > > > > Inspect the back plane for damaged traces and look at the
> > >> capacitors. I
> > >> > > had
> > >> > > > > a few that looked questionable. So far, I have replaced all
> the
> > >> big
> > >> > > caps
> > >> > > on
> > >> > > > > the pre-regulator and regulator boards. My feeling is that
> once
> > >> these
> > >> > > go
> > >> > > > > online, they should run as long as possible between repairs.
> > >> > > > >
> > >> > > > > The battery charger circuit may need adjusting. I tweaked
> mine and
> > >> it
> > >> > > > > seemed to work fine.
> > >> > > > >
> > >> > > > > Todd
> > >> > > > >
> > >> > > > >
> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > >> > > randyevans2688@gmail.com>
> > >> > > > > wrote:
> > >> > > > >
> > >> > > > > > I received my Fluke 732A today. Just powered it up but it
> needs
> > >> new
> > >> > > > > > batteries. Any suggestions for sources (I haven't opened
> up the
> > >> unit
> > >> > > > > yet -
> > >> > > > > > I want to make sure it works before doing that). Also
> received
> > >> the
> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
> ran
> > >> data
> > >> > > > > dumper
> > >> > > > > > program to get the CAL data from my 3458A. Should be a busy
> > >> weekend.
> > >> > > > > >
> > >> > > > > > Randy
> > >> > > > > > _______________________________________________
> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > > > To unsubscribe, go to
> > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > > > and follow the instructions there.
> > >> > > > > >
> > >> > > > > _______________________________________________
> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > > To unsubscribe, go to
> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > > and follow the instructions there.
> > >> > > > >
> > >> > > > _______________________________________________
> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > To unsubscribe, go to
> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > and follow the instructions there.
> > >> > >
> > >> > > _______________________________________________
> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > To unsubscribe, go to
> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > and follow the instructions there.
> > >> > >
> > >> > _______________________________________________
> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > and follow the instructions there.
> > >>
> > >> _______________________________________________
> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> and follow the instructions there.
> > >>
> > >
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
RE
Randy Evans
Wed, Aug 27, 2014 12:36 PM
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.
Randy
On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2688@gmail.com
wrote:
I am doing multiple 100 measurements simply to characterize the stability
of the 3458A and 732A units I just bought. After about 10 measurement sets
over 2 days I am seeing a variance of about .5 uV for the 10V output, or
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, acbern@gmx.de wrote:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am already
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal)
unless the amb. temperature is really very stable (half a degree already
adds about 0.25ppm at 10v)
thanks
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during
measurements. After the SMPL symbol no longer blinks I hit MATH 2
get a MATH ERR symbol on the display. I tried it a couple of times
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide,
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
set
the MATH to "Statistics" (store high reading/low reading/ and mean
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and PLC
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
got
a variation of 0.159 uVolts using the same 40 reading method above.
would
use this to determine where your problem might exist. Just having
meter
input shorted will point you in the right direction. Meter, cables
732A.
Sorry for the long dissertation. Friends get mad at me for
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads about
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack.
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be
if it is a distance away, or you have to use a shipper like UPS
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have the
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power
and then power is restored the result will be a different 10
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found
plugging in the lead will generate a thermal difference because
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A and
total difference of 0.159 uV over 40 measurements using 100 PLC
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
Randy,
You have two possible choices. It can be configured with 4 x
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.
Randy
On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans <randyevans2688@gmail.com>
wrote:
> I am doing multiple 100 measurements simply to characterize the stability
> of the 3458A and 732A units I just bought. After about 10 measurement sets
> over 2 days I am seeing a variance of about .5 uV for the 10V output, or
> 0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
> that sound reasonable/
>
> Randy
>
>
> On Tue, Aug 26, 2014 at 5:41 PM, <acbern@gmx.de> wrote:
>
>> hi randy,
>>
>> just for curiosity, why doing 100 measurements at nplc 1000. is this to
>> sample a changing value?
>> when i am doing 10 measurements from a stable signal at nplc 100 (only
>> there many subsequent measuremnts with statistics make sense) I am already
>> getting a stanard deviation below 0.1ppm.
>> in a 30 minute test cycle, i would also be concerned about drifts (acal)
>> unless the amb. temperature is really very stable (half a degree already
>> adds about 0.25ppm at 10v)
>>
>> thanks
>>
>>
>>
>>
>> > Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
>> > Von: "Randy Evans" <randyevans2688@gmail.com>
>> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> > Betreff: Re: [volt-nuts] 732A and Prologix received
>> >
>> > Is there any way to tell when the function key routine is complete? In
>> the
>> > case of taking multiple readings using the DEFKEY and MATH function, I
>> > don't see any indication when the routine is complete. In one
>> particular
>> > case, I am taking a 100 readings with NLPC set for 1000 so its a long
>> while
>> > before it's complete, but i have to guess when it's done.
>> >
>> > Thanks,
>> >
>> > Randy
>> >
>> >
>> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
>> > wrote:
>> >
>> > > Bill,
>> > >
>> > > I am trying to figure out the MATH function without much success. I
>> input
>> > > the sequence you said (I looked up the instructions to understand
>> what you
>> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
>> 4;TRIG; and
>> > > it shows up on the display when I input BLUE F1. I hit ENTER and it
>> takes
>> > > the 40 measurements and the MATH symbol shows on the display during
>> the
>> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2
>> and I
>> > > get a MATH ERR symbol on the display. I tried it a couple of times
>> and the
>> > > same result so I am doing something wrong. Is there a better source
>> for
>> > > explaining how to do front panel masurements than the User Guide,
>> which
>> > > seems oriented at programming automatic rather than manual
>> measurements.
>> > >
>> > > Randy
>> > >
>> > >
>> > >
>> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net>
>> wrote:
>> > >
>> > >> Randy:
>> > >>
>> > >> The MATH function is accessible from the keypad. I don't have
>> an IEEE
>> > >> interface right now that works. You can also program the numeric
>> keypad
>> > >> keys to have preprogrammed functions. DEFKEY
>> > >>
>> > >> I have made my own "low thermal" measurement leads from Pomona
>> #4892
>> > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy
>> at the
>> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
>> plans
>> > >> to do custom cables with 16 ga. bare copper wire that I will twist
>> and
>> > >> then
>> > >> put a braided shield over it. I simply cannot find what I want so I
>> will
>> > >> build my own cable. I have done something like this before and it
>> worked
>> > >> fine. When I get a "round toit".
>> > >>
>> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
>> have
>> > >> used
>> > >> in the past to verify my homemade "low thermal" leads as described
>> above.
>> > >> Frankly I cannot see any difference between using the 1756 cables
>> and my
>> > >> homemade cables once I give them a few minutes for the thermals to go
>> > >> away.
>> > >> As far as I can tell and measure the differences, if any, are below
>> 0.1
>> > >> ppm
>> > >> at 10 volts.
>> > >>
>> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
>> are all
>> > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
>> far
>> > >> as
>> > >> the instability of the readings it is hard to determine which is
>> causing
>> > >> the
>> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
>> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
>> does is
>> > >> set
>> > >> the MATH to "Statistics" (store high reading/low reading/ and mean
>> of the
>> > >> readings) in the registers, the number of readings to "40", the
>> trigger to
>> > >> "hold" (which keeps the meter from triggering until I press "ENTER"
>> and
>> > >> then
>> > >> trigger the sequence of 40 readings when I push the "ENTER" button.
>> You
>> > >> can
>> > >> do all of this manually from the keypads but since I use this
>> sequence a
>> > >> lot
>> > >> I have preprogrammed it. This is after I set digits to "8" and PLC
>> to
>> > >> 100.
>> > >> Once those 40 readings are finished then you can access the various
>> MATH
>> > >> statistic registers, using the menu, by entering "MATH" and then a 2
>> for
>> > >> low, a 4 for mean, and 13 for high. Of course you could do all of
>> this
>> > >> through the IEEE also. The 3458A has a very rich set of measurement
>> > >> commands. I am still learning all of them. It depends upon what I
>> am
>> > >> trying to accomplish.
>> > >>
>> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from
>> > >> resistive dividers from the 10 volt, I don't see how they could
>> contribute
>> > >> to the varying readings you are measuring. I think I would put a
>> short on
>> > >> the input of the 3458A and manually set the range to 1 volt and then
>> > >> observe
>> > >> the variations that way without the 732A involved. When I do this I
>> see a
>> > >> variation from low reading to high reading of 0.125 uVolts and then
>> > >> another
>> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the
>> low
>> > >> side
>> > >> of the measurment terminals, GUARD connected doesn't seem to affect
>> the
>> > >> readings. So that is the base noise of the 3458A without the 732A,
>> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
>> output I
>> > >> got
>> > >> a variation of 0.159 uVolts using the same 40 reading method above.
>> I
>> > >> would
>> > >> use this to determine where your problem might exist. Just having
>> the
>> > >> meter
>> > >> input shorted will point you in the right direction. Meter, cables
>> or
>> > >> 732A.
>> > >>
>> > >> Sorry for the long dissertation. Friends get mad at me for
>> being so
>> > >> detailed sometimes.
>> > >>
>> > >> Bill
>> > >>
>> > >> ----- Original Message -----
>> > >> From: "Randy Evans" <randyevans2688@gmail.com>
>> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> > >> Sent: Sunday, August 24, 2014 7:22 AM
>> > >> Subject: Re: [volt-nuts] 732A and Prologix received
>> > >>
>> > >>
>> > >> > Bill,
>> > >> >
>> > >> > I have convinced myself that the problem I an seeing is due to
>> thermals.
>> > >> > If I move the cables (with gold-plated banana plugs) using a small
>> > >> towel
>> > >> > rather than letting my hand touch the plugs, it is much more
>> stable.
>> > >> If I
>> > >> > then hold the banana plug with my hand after the reading has
>> stabilized,
>> > >> > the reading drifts rapidly upward. I am trying to check the
>> stability
>> > >> of
>> > >> > the reading but I haven't figured out the MATH function yet. I
>> assume
>> > >> this
>> > >> > is a programmed function using GPIB only?
>> > >> >
>> > >> > The stability I am seeing by manually recording readings (using
>> NLPC of
>> > >> 100
>> > >> > and 1000) is much greater than what you are measuring on your
>> system.
>> > >> Not
>> > >> > sure how to ascertain if it's the 3458A or the 732. The value of
>> the
>> > >> > readings are very different between the two - the 3458 reads about
>> 50 uV
>> > >> > high on the 10 V output and about 12 uV low on the 1V output.
>> Rather
>> > >> large
>> > >> > differences (this is after an ACAL). I need to find some better
>> cables
>> > >> to
>> > >> > make sure the errors are not due to thermals again.
>> > >> >
>> > >> > Randy
>> > >> >
>> > >> >
>> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
>> wrote:
>> > >> >
>> > >> > > Randy:
>> > >> > >
>> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
>> will fit
>> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
>> LC-R064R5C and
>> > >> > > others
>> > >> > > that are in this size and package. Order from one of the usual
>> > >> electronics
>> > >> > > distributors like Allied, Mouser, Digikey. This is a very common
>> > >> battery
>> > >> > > as
>> > >> > > it is used in a lot of "EXIT" signs so they are lighted when the
>> power
>> > >> goes
>> > >> > > out. I don't see how the 12V 7AH will fit as they are too
>> large. I
>> > >> guess
>> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
>> volt
>> > >> 4
>> > >> AH
>> > >> > > but the terminals are in the wrong place so you will have to
>> "nibble"
>> > >> out
>> > >> > > the aluminum plate that holds them in the 732A battery pack.
>> You have
>> > >> to
>> > >> > > be
>> > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
>> > >> connection
>> > >> > > leads to deal with and connect correctly. I would stick with
>> the 6V
>> > >> 4AH.
>> > >> > > New batteries will last around 12 to 14 hours before the "CAL"
>> light
>> > >> goes
>> > >> > > out when AC power is not applied. So shipping to Cal Lab can be
>> a
>> > >> problem
>> > >> > > if it is a distance away, or you have to use a shipper like UPS
>> or
>> > >> FEDEX
>> > >> > > and
>> > >> > > you ship the night before and then use their "Morning delivery"
>> and
>> > >> the
>> > >> Cal
>> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
>> course
>> > >> you
>> > >> > > could always strap another battery on the 732A and hook it up to
>> the
>> > >> "ext
>> > >> > > power" plug to last longer. I have seen it done. The issue is
>> to get
>> > >> the
>> > >> > > Cal Lab to charge the extra battery before they ship the 732A
>> back to
>> > >> you.
>> > >> > >
>> > >> > > When you remove and work on the battery pack always have the
>> AC
>> > >> power
>> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
>> regulated
>> > >> supply
>> > >> > > is working.
>> > >> > >
>> > >> > > The "CAL" light is to indicate that power has not been lost
>> to the
>> > >> > > Reference Amp or other associated circuits. When the raw supply
>> > >> (battery)
>> > >> > > voltage drops below about 21 volts the "CAL" light will go out.
>> Below
>> > >> that
>> > >> > > voltage the heater circuits will not work correctly and the 18.6
>> volt
>> > >> > > regulated supply will not regulate. The requirement is that the
>> > >> Reference
>> > >> > > Amp be kept "alive" at all times to maintain the output voltage
>> that
>> > >> was
>> > >> > > measured at the time of the most recent Calibration or
>> Certification.
>> > >> When
>> > >> > > the semiconductor junctions are unbiased and cool off when power
>> is
>> > >> lost,
>> > >> > > and then power is restored the result will be a different 10
>> volts
>> > >> than
>> > >> > > before the power failure. My experience is that after all of the
>> > >> years
>> > >> > > that
>> > >> > > these units have been powered up, this won't happen and when
>> power is
>> > >> lost
>> > >> > > and then restored, even months later, the 732A will come back to
>> > >> almost
>> > >> > > exactly the same 10 volts as when they lost power, usually with
>> in 0.2
>> > >> PPM
>> > >> > > after 24 hours of "warm up".
>> > >> > >
>> > >> > > What type of hookup leads are you using when measuring the 1
>> volt
>> > >> > > output? What is the PLC set to? I always use 100 PLC to measure
>> > >> this.
>> > >> If
>> > >> > > you don't have "low thermal" connection leads you can experience
>> uV
>> > >> changes
>> > >> > > for a minute or more after plugging in the leads due to the
>> "thermals"
>> > >> > > generated because of the difference in temperature between the
>> banana
>> > >> jacks
>> > >> > > on the 732A and the banana plugs of the leads. I have found
>> that even
>> > >> just
>> > >> > > plugging in the lead will generate a thermal difference because
>> of
>> > >> > > difference of temps and some heating due to the physical act of
>> just
>> > >> > > inserting the plug because of friction between the jack and plug
>> (my
>> > >> theory
>> > >> > > at any rate). You have to allow at least a minute or more before
>> > >> being
>> > >> > > able
>> > >> > > to make a measurement after plugging in the leads. I just
>> measured
>> > >> the
>> > >> > > variation of the 1 volt output of my 732A and using my 3458A and
>> I
>> > >> got a
>> > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC
>> on
>> > >> the
>> > >> 1
>> > >> > > volt range of the 3458A. Using the MATH function and all of the
>> data
>> > >> you
>> > >> > > can collect. That was after waiting for several minutes after
>> > >> plugging
>> > >> in
>> > >> > > the leads.
>> > >> > >
>> > >> > > I hope all of this helps.
>> > >> > >
>> > >> > > Bill
>> > >> > >
>> > >> > >
>> > >> > >
>> > >> > > ----- Original Message -----
>> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
>> > >> > > To: "Discussion of precise voltage measurement" <
>> volt-nuts@febo.com>
>> > >> > > Sent: Friday, August 22, 2014 9:03 PM
>> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
>> > >> > >
>> > >> > >
>> > >> > > > Todd,
>> > >> > > >
>> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
>> batteries
>> > >> that
>> > >> I
>> > >> > > > keep topped off and they have very low current draw (~2 to 3
>> mA at
>> > >> 13.5
>> > >> > > > VDC) when charged and at their float voltage, so I am pretty
>> sure
>> > >> they
>> > >> > > are
>> > >> > > > in good condition. I will look at getting those in the units
>> after
>> > >> I
>> > >> > > > ascertain the condition of the 732.
>> > >> > > >
>> > >> > > > So now I have a what appears to be a functioning 3458A and a
>> 732A
>> > >> but
>> > >> > > they
>> > >> > > > slightly disagree. I am like the man with two watches that
>> disagree
>> > >> on
>> > >> > > the
>> > >> > > > time - which is correct? For the moment, i am only concerned
>> with
>> > >> > > > stability. The need for absolute accuracy will come later.
>> > >> > > >
>> > >> > > > Randy
>> > >> > > >
>> > >> > > >
>> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
>> tmicallef@gmail.com
>> > >> >
>> > >> > > wrote:
>> > >> > > >
>> > >> > > > > Randy,
>> > >> > > > >
>> > >> > > > > You have two possible choices. It can be configured with 4 x
>> 6v
>> > >> 4Ah
>> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
>> owner
>> > >> has
>> > >> > > > > modified the battery pack already. A couple of mine needed a
>> > >> nibbler
>> > >> > > tool
>> > >> > > > > to remove enough of the aluminum cover that fits over the
>> tops of
>> > >> the
>> > >> > > > > batteries. The original cover will short out to the battery
>> tabs
>> > >> > > regardless
>> > >> > > > > of the battery configuration if this is not done.
>> > >> > > > >
>> > >> > > > > You can find larger capacity batteries that will give you
>> slightly
>> > >> more
>> > >> > > > > battery life. I lost a couple sets of mail-order batteries
>> after a
>> > >> few
>> > >> > > > > extended outages. I would recommend going with locally bought
>> > >> batteries
>> > >> > > > > instead of the cheaper mail order. My local Batteries Plus
>> will
>> > >> > > typically
>> > >> > > > > have some warranty if I remember correctly. Moving forward I
>> will
>> > >> only
>> > >> > > use
>> > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
>> > >> equalize
>> > >> > > them
>> > >> > > > > before putting them in the 732A. I think the cheap batteries
>> did
>> > >> not
>> > >> > > > > discharge equally, and would not recover when power was
>> applied.
>> > >> > > > >
>> > >> > > > > Inspect the back plane for damaged traces and look at the
>> > >> capacitors. I
>> > >> > > had
>> > >> > > > > a few that looked questionable. So far, I have replaced all
>> the
>> > >> big
>> > >> > > caps
>> > >> > > on
>> > >> > > > > the pre-regulator and regulator boards. My feeling is that
>> once
>> > >> these
>> > >> > > go
>> > >> > > > > online, they should run as long as possible between repairs.
>> > >> > > > >
>> > >> > > > > The battery charger circuit may need adjusting. I tweaked
>> mine and
>> > >> it
>> > >> > > > > seemed to work fine.
>> > >> > > > >
>> > >> > > > > Todd
>> > >> > > > >
>> > >> > > > >
>> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
>> > >> > > randyevans2688@gmail.com>
>> > >> > > > > wrote:
>> > >> > > > >
>> > >> > > > > > I received my Fluke 732A today. Just powered it up but it
>> needs
>> > >> new
>> > >> > > > > > batteries. Any suggestions for sources (I haven't opened
>> up the
>> > >> unit
>> > >> > > > > yet -
>> > >> > > > > > I want to make sure it works before doing that). Also
>> received
>> > >> the
>> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
>> ran
>> > >> data
>> > >> > > > > dumper
>> > >> > > > > > program to get the CAL data from my 3458A. Should be a
>> busy
>> > >> weekend.
>> > >> > > > > >
>> > >> > > > > > Randy
>> > >> > > > > > _______________________________________________
>> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > >> > > > > > To unsubscribe, go to
>> > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > >> > > > > > and follow the instructions there.
>> > >> > > > > >
>> > >> > > > > _______________________________________________
>> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > >> > > > > To unsubscribe, go to
>> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > >> > > > > and follow the instructions there.
>> > >> > > > >
>> > >> > > > _______________________________________________
>> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > >> > > > To unsubscribe, go to
>> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > >> > > > and follow the instructions there.
>> > >> > >
>> > >> > > _______________________________________________
>> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > >> > > To unsubscribe, go to
>> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > >> > > and follow the instructions there.
>> > >> > >
>> > >> > _______________________________________________
>> > >> > volt-nuts mailing list -- volt-nuts@febo.com
>> > >> > To unsubscribe, go to
>> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > >> > and follow the instructions there.
>> > >>
>> > >> _______________________________________________
>> > >> volt-nuts mailing list -- volt-nuts@febo.com
>> > >> To unsubscribe, go to
>> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > >> and follow the instructions there.
>> > >>
>> > >
>> > >
>> > _______________________________________________
>> > volt-nuts mailing list -- volt-nuts@febo.com
>> > To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > and follow the instructions there.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
>
A
acbern@gmx.de
Wed, Aug 27, 2014 1:41 PM
well, what you do is to measure the stability of the 3458a. 1ppm drift though sounds much as overall short term average variation if your temperature is stable and your 3458a is always on. probably the temp is not stable, and that is what you see, amongst potentially some other smaller drifts like emf. before every cycle you should do an acal dcv, if you still see 10uv +/- drifts that then seems too much. (one thing I need to add is that one of my 3458a, not yet modified to have lower reference temperature, has drifts when switched on and off. that is not so much the case with modified reference. but my assumption re the above is that your meter is always on, as I said)
732a references et al do have small short term drifts, you can determine them with a josephson element (these guys told me), but the 732a is certainly very stable short term compared to a 3458a. my 732a e.g. has a drift of 0.2ppm per year.
Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.
Randy
On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2688@gmail.com
wrote:
I am doing multiple 100 measurements simply to characterize the stability
of the 3458A and 732A units I just bought. After about 10 measurement sets
over 2 days I am seeing a variance of about .5 uV for the 10V output, or
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, acbern@gmx.de wrote:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am already
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal)
unless the amb. temperature is really very stable (half a degree already
adds about 0.25ppm at 10v)
thanks
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during
measurements. After the SMPL symbol no longer blinks I hit MATH 2
get a MATH ERR symbol on the display. I tried it a couple of times
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide,
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
set
the MATH to "Statistics" (store high reading/low reading/ and mean
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and PLC
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
got
a variation of 0.159 uVolts using the same 40 reading method above.
would
use this to determine where your problem might exist. Just having
meter
input shorted will point you in the right direction. Meter, cables
732A.
Sorry for the long dissertation. Friends get mad at me for
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads about
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack.
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be
if it is a distance away, or you have to use a shipper like UPS
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have the
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power
and then power is restored the result will be a different 10
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found
plugging in the lead will generate a thermal difference because
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A and
total difference of 0.159 uV over 40 measurements using 100 PLC
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
Randy,
You have two possible choices. It can be configured with 4 x
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
well, what you do is to measure the stability of the 3458a. 1ppm drift though sounds much as overall short term average variation if your temperature is stable and your 3458a is always on. probably the temp is not stable, and that is what you see, amongst potentially some other smaller drifts like emf. before every cycle you should do an acal dcv, if you still see 10uv +/- drifts that then seems too much. (one thing I need to add is that one of my 3458a, not yet modified to have lower reference temperature, has drifts when switched on and off. that is not so much the case with modified reference. but my assumption re the above is that your meter is always on, as I said)
732a references et al do have small short term drifts, you can determine them with a josephson element (these guys told me), but the 732a is certainly very stable short term compared to a 3458a. my 732a e.g. has a drift of 0.2ppm per year.
> Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
> Von: "Randy Evans" <randyevans2688@gmail.com>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> I forgot to mention that I reduced the number of measurements to 100 per
> set since I wasn't seeing much difference in the variance between 100 and
> 1000 measurements and the 1000 measurement per set takes too long.
>
> Randy
>
>
> On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > I am doing multiple 100 measurements simply to characterize the stability
> > of the 3458A and 732A units I just bought. After about 10 measurement sets
> > over 2 days I am seeing a variance of about .5 uV for the 10V output, or
> > 0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
> > that sound reasonable/
> >
> > Randy
> >
> >
> > On Tue, Aug 26, 2014 at 5:41 PM, <acbern@gmx.de> wrote:
> >
> >> hi randy,
> >>
> >> just for curiosity, why doing 100 measurements at nplc 1000. is this to
> >> sample a changing value?
> >> when i am doing 10 measurements from a stable signal at nplc 100 (only
> >> there many subsequent measuremnts with statistics make sense) I am already
> >> getting a stanard deviation below 0.1ppm.
> >> in a 30 minute test cycle, i would also be concerned about drifts (acal)
> >> unless the amb. temperature is really very stable (half a degree already
> >> adds about 0.25ppm at 10v)
> >>
> >> thanks
> >>
> >>
> >>
> >>
> >> > Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
> >> > Von: "Randy Evans" <randyevans2688@gmail.com>
> >> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> > Betreff: Re: [volt-nuts] 732A and Prologix received
> >> >
> >> > Is there any way to tell when the function key routine is complete? In
> >> the
> >> > case of taking multiple readings using the DEFKEY and MATH function, I
> >> > don't see any indication when the routine is complete. In one
> >> particular
> >> > case, I am taking a 100 readings with NLPC set for 1000 so its a long
> >> while
> >> > before it's complete, but i have to guess when it's done.
> >> >
> >> > Thanks,
> >> >
> >> > Randy
> >> >
> >> >
> >> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> >> > wrote:
> >> >
> >> > > Bill,
> >> > >
> >> > > I am trying to figure out the MATH function without much success. I
> >> input
> >> > > the sequence you said (I looked up the instructions to understand
> >> what you
> >> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
> >> 4;TRIG; and
> >> > > it shows up on the display when I input BLUE F1. I hit ENTER and it
> >> takes
> >> > > the 40 measurements and the MATH symbol shows on the display during
> >> the
> >> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2
> >> and I
> >> > > get a MATH ERR symbol on the display. I tried it a couple of times
> >> and the
> >> > > same result so I am doing something wrong. Is there a better source
> >> for
> >> > > explaining how to do front panel masurements than the User Guide,
> >> which
> >> > > seems oriented at programming automatic rather than manual
> >> measurements.
> >> > >
> >> > > Randy
> >> > >
> >> > >
> >> > >
> >> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net>
> >> wrote:
> >> > >
> >> > >> Randy:
> >> > >>
> >> > >> The MATH function is accessible from the keypad. I don't have
> >> an IEEE
> >> > >> interface right now that works. You can also program the numeric
> >> keypad
> >> > >> keys to have preprogrammed functions. DEFKEY
> >> > >>
> >> > >> I have made my own "low thermal" measurement leads from Pomona
> >> #4892
> >> > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy
> >> at the
> >> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
> >> plans
> >> > >> to do custom cables with 16 ga. bare copper wire that I will twist
> >> and
> >> > >> then
> >> > >> put a braided shield over it. I simply cannot find what I want so I
> >> will
> >> > >> build my own cable. I have done something like this before and it
> >> worked
> >> > >> fine. When I get a "round toit".
> >> > >>
> >> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
> >> have
> >> > >> used
> >> > >> in the past to verify my homemade "low thermal" leads as described
> >> above.
> >> > >> Frankly I cannot see any difference between using the 1756 cables
> >> and my
> >> > >> homemade cables once I give them a few minutes for the thermals to go
> >> > >> away.
> >> > >> As far as I can tell and measure the differences, if any, are below
> >> 0.1
> >> > >> ppm
> >> > >> at 10 volts.
> >> > >>
> >> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
> >> are all
> >> > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
> >> far
> >> > >> as
> >> > >> the instability of the readings it is hard to determine which is
> >> causing
> >> > >> the
> >> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> >> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
> >> does is
> >> > >> set
> >> > >> the MATH to "Statistics" (store high reading/low reading/ and mean
> >> of the
> >> > >> readings) in the registers, the number of readings to "40", the
> >> trigger to
> >> > >> "hold" (which keeps the meter from triggering until I press "ENTER"
> >> and
> >> > >> then
> >> > >> trigger the sequence of 40 readings when I push the "ENTER" button.
> >> You
> >> > >> can
> >> > >> do all of this manually from the keypads but since I use this
> >> sequence a
> >> > >> lot
> >> > >> I have preprogrammed it. This is after I set digits to "8" and PLC
> >> to
> >> > >> 100.
> >> > >> Once those 40 readings are finished then you can access the various
> >> MATH
> >> > >> statistic registers, using the menu, by entering "MATH" and then a 2
> >> for
> >> > >> low, a 4 for mean, and 13 for high. Of course you could do all of
> >> this
> >> > >> through the IEEE also. The 3458A has a very rich set of measurement
> >> > >> commands. I am still learning all of them. It depends upon what I
> >> am
> >> > >> trying to accomplish.
> >> > >>
> >> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> >> > >> resistive dividers from the 10 volt, I don't see how they could
> >> contribute
> >> > >> to the varying readings you are measuring. I think I would put a
> >> short on
> >> > >> the input of the 3458A and manually set the range to 1 volt and then
> >> > >> observe
> >> > >> the variations that way without the 732A involved. When I do this I
> >> see a
> >> > >> variation from low reading to high reading of 0.125 uVolts and then
> >> > >> another
> >> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the
> >> low
> >> > >> side
> >> > >> of the measurment terminals, GUARD connected doesn't seem to affect
> >> the
> >> > >> readings. So that is the base noise of the 3458A without the 732A,
> >> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
> >> output I
> >> > >> got
> >> > >> a variation of 0.159 uVolts using the same 40 reading method above.
> >> I
> >> > >> would
> >> > >> use this to determine where your problem might exist. Just having
> >> the
> >> > >> meter
> >> > >> input shorted will point you in the right direction. Meter, cables
> >> or
> >> > >> 732A.
> >> > >>
> >> > >> Sorry for the long dissertation. Friends get mad at me for
> >> being so
> >> > >> detailed sometimes.
> >> > >>
> >> > >> Bill
> >> > >>
> >> > >> ----- Original Message -----
> >> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> >> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> > >> Sent: Sunday, August 24, 2014 7:22 AM
> >> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >>
> >> > >>
> >> > >> > Bill,
> >> > >> >
> >> > >> > I have convinced myself that the problem I an seeing is due to
> >> thermals.
> >> > >> > If I move the cables (with gold-plated banana plugs) using a small
> >> > >> towel
> >> > >> > rather than letting my hand touch the plugs, it is much more
> >> stable.
> >> > >> If I
> >> > >> > then hold the banana plug with my hand after the reading has
> >> stabilized,
> >> > >> > the reading drifts rapidly upward. I am trying to check the
> >> stability
> >> > >> of
> >> > >> > the reading but I haven't figured out the MATH function yet. I
> >> assume
> >> > >> this
> >> > >> > is a programmed function using GPIB only?
> >> > >> >
> >> > >> > The stability I am seeing by manually recording readings (using
> >> NLPC of
> >> > >> 100
> >> > >> > and 1000) is much greater than what you are measuring on your
> >> system.
> >> > >> Not
> >> > >> > sure how to ascertain if it's the 3458A or the 732. The value of
> >> the
> >> > >> > readings are very different between the two - the 3458 reads about
> >> 50 uV
> >> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> >> Rather
> >> > >> large
> >> > >> > differences (this is after an ACAL). I need to find some better
> >> cables
> >> > >> to
> >> > >> > make sure the errors are not due to thermals again.
> >> > >> >
> >> > >> > Randy
> >> > >> >
> >> > >> >
> >> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
> >> wrote:
> >> > >> >
> >> > >> > > Randy:
> >> > >> > >
> >> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
> >> will fit
> >> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
> >> LC-R064R5C and
> >> > >> > > others
> >> > >> > > that are in this size and package. Order from one of the usual
> >> > >> electronics
> >> > >> > > distributors like Allied, Mouser, Digikey. This is a very common
> >> > >> battery
> >> > >> > > as
> >> > >> > > it is used in a lot of "EXIT" signs so they are lighted when the
> >> power
> >> > >> goes
> >> > >> > > out. I don't see how the 12V 7AH will fit as they are too
> >> large. I
> >> > >> guess
> >> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
> >> volt
> >> > >> 4
> >> > >> AH
> >> > >> > > but the terminals are in the wrong place so you will have to
> >> "nibble"
> >> > >> out
> >> > >> > > the aluminum plate that holds them in the 732A battery pack.
> >> You have
> >> > >> to
> >> > >> > > be
> >> > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> >> > >> connection
> >> > >> > > leads to deal with and connect correctly. I would stick with
> >> the 6V
> >> > >> 4AH.
> >> > >> > > New batteries will last around 12 to 14 hours before the "CAL"
> >> light
> >> > >> goes
> >> > >> > > out when AC power is not applied. So shipping to Cal Lab can be
> >> a
> >> > >> problem
> >> > >> > > if it is a distance away, or you have to use a shipper like UPS
> >> or
> >> > >> FEDEX
> >> > >> > > and
> >> > >> > > you ship the night before and then use their "Morning delivery"
> >> and
> >> > >> the
> >> > >> Cal
> >> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> >> course
> >> > >> you
> >> > >> > > could always strap another battery on the 732A and hook it up to
> >> the
> >> > >> "ext
> >> > >> > > power" plug to last longer. I have seen it done. The issue is
> >> to get
> >> > >> the
> >> > >> > > Cal Lab to charge the extra battery before they ship the 732A
> >> back to
> >> > >> you.
> >> > >> > >
> >> > >> > > When you remove and work on the battery pack always have the
> >> AC
> >> > >> power
> >> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> >> regulated
> >> > >> supply
> >> > >> > > is working.
> >> > >> > >
> >> > >> > > The "CAL" light is to indicate that power has not been lost
> >> to the
> >> > >> > > Reference Amp or other associated circuits. When the raw supply
> >> > >> (battery)
> >> > >> > > voltage drops below about 21 volts the "CAL" light will go out.
> >> Below
> >> > >> that
> >> > >> > > voltage the heater circuits will not work correctly and the 18.6
> >> volt
> >> > >> > > regulated supply will not regulate. The requirement is that the
> >> > >> Reference
> >> > >> > > Amp be kept "alive" at all times to maintain the output voltage
> >> that
> >> > >> was
> >> > >> > > measured at the time of the most recent Calibration or
> >> Certification.
> >> > >> When
> >> > >> > > the semiconductor junctions are unbiased and cool off when power
> >> is
> >> > >> lost,
> >> > >> > > and then power is restored the result will be a different 10
> >> volts
> >> > >> than
> >> > >> > > before the power failure. My experience is that after all of the
> >> > >> years
> >> > >> > > that
> >> > >> > > these units have been powered up, this won't happen and when
> >> power is
> >> > >> lost
> >> > >> > > and then restored, even months later, the 732A will come back to
> >> > >> almost
> >> > >> > > exactly the same 10 volts as when they lost power, usually with
> >> in 0.2
> >> > >> PPM
> >> > >> > > after 24 hours of "warm up".
> >> > >> > >
> >> > >> > > What type of hookup leads are you using when measuring the 1
> >> volt
> >> > >> > > output? What is the PLC set to? I always use 100 PLC to measure
> >> > >> this.
> >> > >> If
> >> > >> > > you don't have "low thermal" connection leads you can experience
> >> uV
> >> > >> changes
> >> > >> > > for a minute or more after plugging in the leads due to the
> >> "thermals"
> >> > >> > > generated because of the difference in temperature between the
> >> banana
> >> > >> jacks
> >> > >> > > on the 732A and the banana plugs of the leads. I have found
> >> that even
> >> > >> just
> >> > >> > > plugging in the lead will generate a thermal difference because
> >> of
> >> > >> > > difference of temps and some heating due to the physical act of
> >> just
> >> > >> > > inserting the plug because of friction between the jack and plug
> >> (my
> >> > >> theory
> >> > >> > > at any rate). You have to allow at least a minute or more before
> >> > >> being
> >> > >> > > able
> >> > >> > > to make a measurement after plugging in the leads. I just
> >> measured
> >> > >> the
> >> > >> > > variation of the 1 volt output of my 732A and using my 3458A and
> >> I
> >> > >> got a
> >> > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC
> >> on
> >> > >> the
> >> > >> 1
> >> > >> > > volt range of the 3458A. Using the MATH function and all of the
> >> data
> >> > >> you
> >> > >> > > can collect. That was after waiting for several minutes after
> >> > >> plugging
> >> > >> in
> >> > >> > > the leads.
> >> > >> > >
> >> > >> > > I hope all of this helps.
> >> > >> > >
> >> > >> > > Bill
> >> > >> > >
> >> > >> > >
> >> > >> > >
> >> > >> > > ----- Original Message -----
> >> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> >> > >> > > To: "Discussion of precise voltage measurement" <
> >> volt-nuts@febo.com>
> >> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> >> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >> > >
> >> > >> > >
> >> > >> > > > Todd,
> >> > >> > > >
> >> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> >> batteries
> >> > >> that
> >> > >> I
> >> > >> > > > keep topped off and they have very low current draw (~2 to 3
> >> mA at
> >> > >> 13.5
> >> > >> > > > VDC) when charged and at their float voltage, so I am pretty
> >> sure
> >> > >> they
> >> > >> > > are
> >> > >> > > > in good condition. I will look at getting those in the units
> >> after
> >> > >> I
> >> > >> > > > ascertain the condition of the 732.
> >> > >> > > >
> >> > >> > > > So now I have a what appears to be a functioning 3458A and a
> >> 732A
> >> > >> but
> >> > >> > > they
> >> > >> > > > slightly disagree. I am like the man with two watches that
> >> disagree
> >> > >> on
> >> > >> > > the
> >> > >> > > > time - which is correct? For the moment, i am only concerned
> >> with
> >> > >> > > > stability. The need for absolute accuracy will come later.
> >> > >> > > >
> >> > >> > > > Randy
> >> > >> > > >
> >> > >> > > >
> >> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
> >> tmicallef@gmail.com
> >> > >> >
> >> > >> > > wrote:
> >> > >> > > >
> >> > >> > > > > Randy,
> >> > >> > > > >
> >> > >> > > > > You have two possible choices. It can be configured with 4 x
> >> 6v
> >> > >> 4Ah
> >> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> >> owner
> >> > >> has
> >> > >> > > > > modified the battery pack already. A couple of mine needed a
> >> > >> nibbler
> >> > >> > > tool
> >> > >> > > > > to remove enough of the aluminum cover that fits over the
> >> tops of
> >> > >> the
> >> > >> > > > > batteries. The original cover will short out to the battery
> >> tabs
> >> > >> > > regardless
> >> > >> > > > > of the battery configuration if this is not done.
> >> > >> > > > >
> >> > >> > > > > You can find larger capacity batteries that will give you
> >> slightly
> >> > >> more
> >> > >> > > > > battery life. I lost a couple sets of mail-order batteries
> >> after a
> >> > >> few
> >> > >> > > > > extended outages. I would recommend going with locally bought
> >> > >> batteries
> >> > >> > > > > instead of the cheaper mail order. My local Batteries Plus
> >> will
> >> > >> > > typically
> >> > >> > > > > have some warranty if I remember correctly. Moving forward I
> >> will
> >> > >> only
> >> > >> > > use
> >> > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> >> > >> equalize
> >> > >> > > them
> >> > >> > > > > before putting them in the 732A. I think the cheap batteries
> >> did
> >> > >> not
> >> > >> > > > > discharge equally, and would not recover when power was
> >> applied.
> >> > >> > > > >
> >> > >> > > > > Inspect the back plane for damaged traces and look at the
> >> > >> capacitors. I
> >> > >> > > had
> >> > >> > > > > a few that looked questionable. So far, I have replaced all
> >> the
> >> > >> big
> >> > >> > > caps
> >> > >> > > on
> >> > >> > > > > the pre-regulator and regulator boards. My feeling is that
> >> once
> >> > >> these
> >> > >> > > go
> >> > >> > > > > online, they should run as long as possible between repairs.
> >> > >> > > > >
> >> > >> > > > > The battery charger circuit may need adjusting. I tweaked
> >> mine and
> >> > >> it
> >> > >> > > > > seemed to work fine.
> >> > >> > > > >
> >> > >> > > > > Todd
> >> > >> > > > >
> >> > >> > > > >
> >> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> >> > >> > > randyevans2688@gmail.com>
> >> > >> > > > > wrote:
> >> > >> > > > >
> >> > >> > > > > > I received my Fluke 732A today. Just powered it up but it
> >> needs
> >> > >> new
> >> > >> > > > > > batteries. Any suggestions for sources (I haven't opened
> >> up the
> >> > >> unit
> >> > >> > > > > yet -
> >> > >> > > > > > I want to make sure it works before doing that). Also
> >> received
> >> > >> the
> >> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
> >> ran
> >> > >> data
> >> > >> > > > > dumper
> >> > >> > > > > > program to get the CAL data from my 3458A. Should be a
> >> busy
> >> > >> weekend.
> >> > >> > > > > >
> >> > >> > > > > > Randy
> >> > >> > > > > > _______________________________________________
> >> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > > > > To unsubscribe, go to
> >> > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > > > > and follow the instructions there.
> >> > >> > > > > >
> >> > >> > > > > _______________________________________________
> >> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > > > To unsubscribe, go to
> >> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > > > and follow the instructions there.
> >> > >> > > > >
> >> > >> > > > _______________________________________________
> >> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > > To unsubscribe, go to
> >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > > and follow the instructions there.
> >> > >> > >
> >> > >> > > _______________________________________________
> >> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > To unsubscribe, go to
> >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > and follow the instructions there.
> >> > >> > >
> >> > >> > _______________________________________________
> >> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > To unsubscribe, go to
> >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > and follow the instructions there.
> >> > >>
> >> > >> _______________________________________________
> >> > >> volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> To unsubscribe, go to
> >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> and follow the instructions there.
> >> > >>
> >> > >
> >> > >
> >> > _______________________________________________
> >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > and follow the instructions there.
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >>
> >
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
BG
Bill Gold
Wed, Aug 27, 2014 2:30 PM
Randy:
Something I had never tried to measure. As I have found out in the past
there is a lot of overhead going on in the meter during and after a
measurement.
In thinking about this I turned "OFF" the autozero "AZERO" and the time
for each "SMPL" was cut in half to around the estimated 16.66 seconds for
1000 PLC. So it becomes obvious that the meter makes an autozero
measurement for 1000 PLC and then the actual measurement for 1000 PLC which
explains the 33 seconds. Makes sense. This is probably why the AZERO menu
gives you ON, OFF and ONCE. For short measurement sequences you just
autozero ONCE at the start.
Thanks for the observation, this helps me. Everyday I learn something
new.
Bill
Thanks for the information
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Tuesday, August 26, 2014 2:16 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.
Randy
On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3637@att.net wrote:
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds
per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete?
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during
measurements. After the SMPL symbol no longer blinks I hit MATH 2
get a MATH ERR symbol on the display. I tried it a couple of times
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide,
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy
time. It is tin plated copper, shielded twisted pair 20 ga. I
to do custom cables with 16 ga. bare copper wire that I will twist
then
put a braided shield over it. I simply cannot find what I want so
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables
homemade cables once I give them a few minutes for the thermals to
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
adjustable you may be seeing a misadjusted 1 volt from the 732A.
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
set
the MATH to "Statistics" (store high reading/low reading/ and mean
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and PLC
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of
commands. I am still learning all of them. It depends upon what I
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and
observe
the variations that way without the 732A involved. When I do this
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
got
a variation of 0.159 uVolts using the same 40 reading method above.
would
use this to determine where your problem might exist. Just having
meter
input shorted will point you in the right direction. Meter, cables
732A.
Sorry for the long dissertation. Friends get mad at me for
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a
rather than letting my hand touch the plugs, it is much more
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very
as
it is used in a lot of "EXIT" signs so they are lighted when
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2 x
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack.
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can
if it is a distance away, or you have to use a shipper like UPS
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up
power" plug to last longer. I have seen it done. The issue is
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost
Reference Amp or other associated circuits. When the raw
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the
regulated supply will not regulate. The requirement is that
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when
and then power is restored the result will be a different 10
before the power failure. My experience is that after all of
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come back
exactly the same 10 volts as when they lost power, usually with
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the
output? What is the PLC set to? I always use 100 PLC to
you don't have "low thermal" connection leads you can
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found
plugging in the lead will generate a thermal difference because
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and
at any rate). You have to allow at least a minute or more
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A
total difference of 0.159 uV over 40 measurements using 100
volt range of the 3458A. Using the MATH function and all of
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but
batteries. Any suggestions for sources (I haven't opened
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims'
program to get the CAL data from my 3458A. Should be a
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Randy:
Something I had never tried to measure. As I have found out in the past
there is a lot of overhead going on in the meter during and after a
measurement.
In thinking about this I turned "OFF" the autozero "AZERO" and the time
for each "SMPL" was cut in half to around the estimated 16.66 seconds for
1000 PLC. So it becomes obvious that the meter makes an autozero
measurement for 1000 PLC and then the actual measurement for 1000 PLC which
explains the 33 seconds. Makes sense. This is probably why the AZERO menu
gives you ON, OFF and ONCE. For short measurement sequences you just
autozero ONCE at the start.
Thanks for the observation, this helps me. Everyday I learn something
new.
Bill
Thanks for the information
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Tuesday, August 26, 2014 2:16 PM
Subject: Re: [volt-nuts] 732A and Prologix received
> Bill,
>
> I measured the time between SMPL symbols with NPLC set to 1000 and it is
> approximately 33 seconds. It takes an hour to complete 100 readings.
>
> Randy
>
>
> On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold <wpgold3637@att.net> wrote:
>
> > Not that I know of. Just wait and when you don't see a "SMPL" on the
> > display it is done. But then with 1000 PLC that is around 16.66 seconds
> > per
> > reading times 100 readings is somewhere around 28 minutes and there is
> > probably some overhead time so around 30 minutes. Not from the front
panel
> > at any rate.
> >
> > Bill
> >
> > ----- Original Message -----
> > From: "Randy Evans" <randyevans2688@gmail.com>
> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Sent: Monday, August 25, 2014 9:23 PM
> > Subject: Re: [volt-nuts] 732A and Prologix received
> >
> >
> > > Is there any way to tell when the function key routine is complete?
In
> > the
> > > case of taking multiple readings using the DEFKEY and MATH function, I
> > > don't see any indication when the routine is complete. In one
particular
> > > case, I am taking a 100 readings with NLPC set for 1000 so its a long
> > while
> > > before it's complete, but i have to guess when it's done.
> > >
> > > Thanks,
> > >
> > > Randy
> > >
> > >
> > > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans
<randyevans2688@gmail.com>
> > > wrote:
> > >
> > > > Bill,
> > > >
> > > > I am trying to figure out the MATH function without much success. I
> > input
> > > > the sequence you said (I looked up the instructions to understand
what
> > you
> > > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
4;TRIG;
> > and
> > > > it shows up on the display when I input BLUE F1. I hit ENTER and it
> > takes
> > > > the 40 measurements and the MATH symbol shows on the display during
the
> > > > measurements. After the SMPL symbol no longer blinks I hit MATH 2
and
> > I
> > > > get a MATH ERR symbol on the display. I tried it a couple of times
and
> > the
> > > > same result so I am doing something wrong. Is there a better source
> > for
> > > > explaining how to do front panel masurements than the User Guide,
which
> > > > seems oriented at programming automatic rather than manual
> > measurements.
> > > >
> > > > Randy
> > > >
> > > >
> > > >
> > > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net>
wrote:
> > > >
> > > >> Randy:
> > > >>
> > > >> The MATH function is accessible from the keypad. I don't have
an
> > IEEE
> > > >> interface right now that works. You can also program the numeric
> > keypad
> > > >> keys to have preprogrammed functions. DEFKEY
> > > >>
> > > >> I have made my own "low thermal" measurement leads from Pomona
> > #4892
> > > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy
at
> > the
> > > >> time. It is tin plated copper, shielded twisted pair 20 ga. I
have
> > plans
> > > >> to do custom cables with 16 ga. bare copper wire that I will twist
and
> > > >> then
> > > >> put a braided shield over it. I simply cannot find what I want so
I
> > will
> > > >> build my own cable. I have done something like this before and it
> > worked
> > > >> fine. When I get a "round toit".
> > > >>
> > > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
> > have
> > > >> used
> > > >> in the past to verify my homemade "low thermal" leads as described
> > above.
> > > >> Frankly I cannot see any difference between using the 1756 cables
and
> > my
> > > >> homemade cables once I give them a few minutes for the thermals to
go
> > > >> away.
> > > >> As far as I can tell and measure the differences, if any, are below
> > 0.1
> > > >> ppm
> > > >> at 10 volts.
> > > >>
> > > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
are
> > all
> > > >> adjustable you may be seeing a misadjusted 1 volt from the 732A.
As
> > far
> > > >> as
> > > >> the instability of the readings it is hard to determine which is
> > causing
> > > >> the
> > > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with
the
> > > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
does
> > is
> > > >> set
> > > >> the MATH to "Statistics" (store high reading/low reading/ and mean
of
> > the
> > > >> readings) in the registers, the number of readings to "40", the
> > trigger
> > to
> > > >> "hold" (which keeps the meter from triggering until I press "ENTER"
> > and
> > > >> then
> > > >> trigger the sequence of 40 readings when I push the "ENTER" button.
> > You
> > > >> can
> > > >> do all of this manually from the keypads but since I use this
sequence
> > a
> > > >> lot
> > > >> I have preprogrammed it. This is after I set digits to "8" and PLC
to
> > > >> 100.
> > > >> Once those 40 readings are finished then you can access the various
> > MATH
> > > >> statistic registers, using the menu, by entering "MATH" and then a
2
> > for
> > > >> low, a 4 for mean, and 13 for high. Of course you could do all of
> > this
> > > >> through the IEEE also. The 3458A has a very rich set of
measurement
> > > >> commands. I am still learning all of them. It depends upon what I
am
> > > >> trying to accomplish.
> > > >>
> > > >> Since the 1.018 and 1.0 volt outputs are passive and derived
from
> > > >> resistive dividers from the 10 volt, I don't see how they could
> > contribute
> > > >> to the varying readings you are measuring. I think I would put a
> > short
> > on
> > > >> the input of the 3458A and manually set the range to 1 volt and
then
> > > >> observe
> > > >> the variations that way without the 732A involved. When I do this
I
> > see a
> > > >> variation from low reading to high reading of 0.125 uVolts and then
> > > >> another
> > > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the
low
> > > >> side
> > > >> of the measurment terminals, GUARD connected doesn't seem to affect
> > the
> > > >> readings. So that is the base noise of the 3458A without the 732A,
> > > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
output
> > I
> > > >> got
> > > >> a variation of 0.159 uVolts using the same 40 reading method above.
I
> > > >> would
> > > >> use this to determine where your problem might exist. Just having
the
> > > >> meter
> > > >> input shorted will point you in the right direction. Meter, cables
or
> > > >> 732A.
> > > >>
> > > >> Sorry for the long dissertation. Friends get mad at me for
being
> > so
> > > >> detailed sometimes.
> > > >>
> > > >> Bill
> > > >>
> > > >> ----- Original Message -----
> > > >> From: "Randy Evans" <randyevans2688@gmail.com>
> > > >> To: "Discussion of precise voltage measurement"
<volt-nuts@febo.com>
> > > >> Sent: Sunday, August 24, 2014 7:22 AM
> > > >> Subject: Re: [volt-nuts] 732A and Prologix received
> > > >>
> > > >>
> > > >> > Bill,
> > > >> >
> > > >> > I have convinced myself that the problem I an seeing is due to
> > thermals.
> > > >> > If I move the cables (with gold-plated banana plugs) using a
small
> > > >> towel
> > > >> > rather than letting my hand touch the plugs, it is much more
stable.
> > > >> If I
> > > >> > then hold the banana plug with my hand after the reading has
> > stabilized,
> > > >> > the reading drifts rapidly upward. I am trying to check the
> > stability
> > > >> of
> > > >> > the reading but I haven't figured out the MATH function yet. I
> > assume
> > > >> this
> > > >> > is a programmed function using GPIB only?
> > > >> >
> > > >> > The stability I am seeing by manually recording readings (using
NLPC
> > of
> > > >> 100
> > > >> > and 1000) is much greater than what you are measuring on your
> > system.
> > > >> Not
> > > >> > sure how to ascertain if it's the 3458A or the 732. The value of
> > the
> > > >> > readings are very different between the two - the 3458 reads
about
> > 50
> > uV
> > > >> > high on the 10 V output and about 12 uV low on the 1V output.
> > Rather
> > > >> large
> > > >> > differences (this is after an ACAL). I need to find some better
> > cables
> > > >> to
> > > >> > make sure the errors are not due to thermals again.
> > > >> >
> > > >> > Randy
> > > >> >
> > > >> >
> > > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
> > wrote:
> > > >> >
> > > >> > > Randy:
> > > >> > >
> > > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
will
> > fit
> > > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
LC-R064R5C
> > and
> > > >> > > others
> > > >> > > that are in this size and package. Order from one of the usual
> > > >> electronics
> > > >> > > distributors like Allied, Mouser, Digikey. This is a very
common
> > > >> battery
> > > >> > > as
> > > >> > > it is used in a lot of "EXIT" signs so they are lighted when
the
> > power
> > > >> goes
> > > >> > > out. I don't see how the 12V 7AH will fit as they are too
large.
> > I
> > > >> guess
> > > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x
6
> > volt
> > > >> 4
> > > >> AH
> > > >> > > but the terminals are in the wrong place so you will have to
> > "nibble"
> > > >> out
> > > >> > > the aluminum plate that holds them in the 732A battery pack.
You
> > have
> > > >> to
> > > >> > > be
> > > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> > > >> connection
> > > >> > > leads to deal with and connect correctly. I would stick with
the
> > 6V
> > > >> 4AH.
> > > >> > > New batteries will last around 12 to 14 hours before the "CAL"
> > light
> > > >> goes
> > > >> > > out when AC power is not applied. So shipping to Cal Lab can
be a
> > > >> problem
> > > >> > > if it is a distance away, or you have to use a shipper like UPS
or
> > > >> FEDEX
> > > >> > > and
> > > >> > > you ship the night before and then use their "Morning delivery"
> > and
> > > >> the
> > > >> Cal
> > > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> > course
> > > >> you
> > > >> > > could always strap another battery on the 732A and hook it up
to
> > the
> > > >> "ext
> > > >> > > power" plug to last longer. I have seen it done. The issue is
to
> > get
> > > >> the
> > > >> > > Cal Lab to charge the extra battery before they ship the 732A
back
> > to
> > > >> you.
> > > >> > >
> > > >> > > When you remove and work on the battery pack always have
the
> > AC
> > > >> power
> > > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> > regulated
> > > >> supply
> > > >> > > is working.
> > > >> > >
> > > >> > > The "CAL" light is to indicate that power has not been lost
to
> > the
> > > >> > > Reference Amp or other associated circuits. When the raw
supply
> > > >> (battery)
> > > >> > > voltage drops below about 21 volts the "CAL" light will go out.
> > Below
> > > >> that
> > > >> > > voltage the heater circuits will not work correctly and the
18.6
> > volt
> > > >> > > regulated supply will not regulate. The requirement is that
the
> > > >> Reference
> > > >> > > Amp be kept "alive" at all times to maintain the output voltage
> > that
> > > >> was
> > > >> > > measured at the time of the most recent Calibration or
> > Certification.
> > > >> When
> > > >> > > the semiconductor junctions are unbiased and cool off when
power
> > is
> > > >> lost,
> > > >> > > and then power is restored the result will be a different 10
volts
> > > >> than
> > > >> > > before the power failure. My experience is that after all of
the
> > > >> years
> > > >> > > that
> > > >> > > these units have been powered up, this won't happen and when
power
> > is
> > > >> lost
> > > >> > > and then restored, even months later, the 732A will come back
to
> > > >> almost
> > > >> > > exactly the same 10 volts as when they lost power, usually with
in
> > 0.2
> > > >> PPM
> > > >> > > after 24 hours of "warm up".
> > > >> > >
> > > >> > > What type of hookup leads are you using when measuring the
1
> > volt
> > > >> > > output? What is the PLC set to? I always use 100 PLC to
measure
> > > >> this.
> > > >> If
> > > >> > > you don't have "low thermal" connection leads you can
experience
> > uV
> > > >> changes
> > > >> > > for a minute or more after plugging in the leads due to the
> > "thermals"
> > > >> > > generated because of the difference in temperature between the
> > banana
> > > >> jacks
> > > >> > > on the 732A and the banana plugs of the leads. I have found
that
> > even
> > > >> just
> > > >> > > plugging in the lead will generate a thermal difference because
of
> > > >> > > difference of temps and some heating due to the physical act of
> > just
> > > >> > > inserting the plug because of friction between the jack and
plug
> > (my
> > > >> theory
> > > >> > > at any rate). You have to allow at least a minute or more
before
> > > >> being
> > > >> > > able
> > > >> > > to make a measurement after plugging in the leads. I just
> > measured
> > > >> the
> > > >> > > variation of the 1 volt output of my 732A and using my 3458A
and I
> > > >> got a
> > > >> > > total difference of 0.159 uV over 40 measurements using 100
PLC
> > on
> > > >> the
> > > >> 1
> > > >> > > volt range of the 3458A. Using the MATH function and all of
the
> > data
> > > >> you
> > > >> > > can collect. That was after waiting for several minutes after
> > > >> plugging
> > > >> in
> > > >> > > the leads.
> > > >> > >
> > > >> > > I hope all of this helps.
> > > >> > >
> > > >> > > Bill
> > > >> > >
> > > >> > >
> > > >> > >
> > > >> > > ----- Original Message -----
> > > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > > >> > > To: "Discussion of precise voltage measurement"
> > <volt-nuts@febo.com>
> > > >> > > Sent: Friday, August 22, 2014 9:03 PM
> > > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > > >> > >
> > > >> > >
> > > >> > > > Todd,
> > > >> > > >
> > > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> > batteries
> > > >> that
> > > >> I
> > > >> > > > keep topped off and they have very low current draw (~2 to 3
mA
> > at
> > > >> 13.5
> > > >> > > > VDC) when charged and at their float voltage, so I am pretty
> > sure
> > > >> they
> > > >> > > are
> > > >> > > > in good condition. I will look at getting those in the units
> > after
> > > >> I
> > > >> > > > ascertain the condition of the 732.
> > > >> > > >
> > > >> > > > So now I have a what appears to be a functioning 3458A and a
> > 732A
> > > >> but
> > > >> > > they
> > > >> > > > slightly disagree. I am like the man with two watches that
> > disagree
> > > >> on
> > > >> > > the
> > > >> > > > time - which is correct? For the moment, i am only
concerned
> > with
> > > >> > > > stability. The need for absolute accuracy will come later.
> > > >> > > >
> > > >> > > > Randy
> > > >> > > >
> > > >> > > >
> > > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
> > <tmicallef@gmail.com
> > > >> >
> > > >> > > wrote:
> > > >> > > >
> > > >> > > > > Randy,
> > > >> > > > >
> > > >> > > > > You have two possible choices. It can be configured with 4
x
> > 6v
> > > >> 4Ah
> > > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> > owner
> > > >> has
> > > >> > > > > modified the battery pack already. A couple of mine needed
a
> > > >> nibbler
> > > >> > > tool
> > > >> > > > > to remove enough of the aluminum cover that fits over the
tops
> > of
> > > >> the
> > > >> > > > > batteries. The original cover will short out to the battery
> > tabs
> > > >> > > regardless
> > > >> > > > > of the battery configuration if this is not done.
> > > >> > > > >
> > > >> > > > > You can find larger capacity batteries that will give you
> > slightly
> > > >> more
> > > >> > > > > battery life. I lost a couple sets of mail-order batteries
> > after a
> > > >> few
> > > >> > > > > extended outages. I would recommend going with locally
bought
> > > >> batteries
> > > >> > > > > instead of the cheaper mail order. My local Batteries Plus
> > will
> > > >> > > typically
> > > >> > > > > have some warranty if I remember correctly. Moving forward
I
> > will
> > > >> only
> > > >> > > use
> > > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> > > >> equalize
> > > >> > > them
> > > >> > > > > before putting them in the 732A. I think the cheap
batteries
> > did
> > > >> not
> > > >> > > > > discharge equally, and would not recover when power was
> > applied.
> > > >> > > > >
> > > >> > > > > Inspect the back plane for damaged traces and look at the
> > > >> capacitors. I
> > > >> > > had
> > > >> > > > > a few that looked questionable. So far, I have replaced all
> > the
> > > >> big
> > > >> > > caps
> > > >> > > on
> > > >> > > > > the pre-regulator and regulator boards. My feeling is that
> > once
> > > >> these
> > > >> > > go
> > > >> > > > > online, they should run as long as possible between
repairs.
> > > >> > > > >
> > > >> > > > > The battery charger circuit may need adjusting. I tweaked
mine
> > and
> > > >> it
> > > >> > > > > seemed to work fine.
> > > >> > > > >
> > > >> > > > > Todd
> > > >> > > > >
> > > >> > > > >
> > > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > >> > > randyevans2688@gmail.com>
> > > >> > > > > wrote:
> > > >> > > > >
> > > >> > > > > > I received my Fluke 732A today. Just powered it up but
it
> > needs
> > > >> new
> > > >> > > > > > batteries. Any suggestions for sources (I haven't opened
up
> > the
> > > >> unit
> > > >> > > > > yet -
> > > >> > > > > > I want to make sure it works before doing that). Also
> > received
> > > >> the
> > > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims'
CAL
> > ran
> > > >> data
> > > >> > > > > dumper
> > > >> > > > > > program to get the CAL data from my 3458A. Should be a
busy
> > > >> weekend.
> > > >> > > > > >
> > > >> > > > > > Randy
> > > >> > > > > > _______________________________________________
> > > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > > To unsubscribe, go to
> > > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> > > > > > and follow the instructions there.
> > > >> > > > > >
> > > >> > > > > _______________________________________________
> > > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > To unsubscribe, go to
> > > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> > > > > and follow the instructions there.
> > > >> > > > >
> > > >> > > > _______________________________________________
> > > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > To unsubscribe, go to
> > > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> > > > and follow the instructions there.
> > > >> > >
> > > >> > > _______________________________________________
> > > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > To unsubscribe, go to
> > > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> > > and follow the instructions there.
> > > >> > >
> > > >> > _______________________________________________
> > > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > To unsubscribe, go to
> > > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> > and follow the instructions there.
> > > >>
> > > >> _______________________________________________
> > > >> volt-nuts mailing list -- volt-nuts@febo.com
> > > >> To unsubscribe, go to
> > > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> and follow the instructions there.
> > > >>
> > > >
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
D
Don@True-Cal
Wed, Aug 27, 2014 2:40 PM
I have included a link to an ~30-hour measurement run that I consistently do that should give you some idea of expected measurement drift over time. Virtually all of the measurement drift is due to the 3458A internal temperature differences around the 7V reference caused by ambient temperature change. The drift associated with the 732A is probably about 2-magnitudes less at this ambient temp drift. In my situation, the inferred ambient temperature is cycling with the home air-conditioning. Note the initial calibration temperatures as well as the ACal temperatures and where the 3458A is measuring exactly 10V relative to those temperatures. My primary 732A has been powered without loss for 4 years and >5 years before that. The 3458A is Agilent (Loveland) Cal'ed yearly and is powered 24-7-365 except for the occasional mains power loss. The graphical measurements is using a homegrown Agilent VEE program. It is very helpful if not essential to get an HPIB interface setup so you can do long term graphical analysis.
The other link was done in winter time when the lab a few degrees cooler.
https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0
https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%203458A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0
I hope these Dropbox links work internationally - I'm new to using this sharing method.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Wednesday, August 27, 2014 8:42 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
well, what you do is to measure the stability of the 3458a. 1ppm drift though sounds much as overall short term average variation if your temperature is stable and your 3458a is always on. probably the temp is not stable, and that is what you see, amongst potentially some other smaller drifts like emf. before every cycle you should do an acal dcv, if you still see 10uv +/- drifts that then seems too much. (one thing I need to add is that one of my 3458a, not yet modified to have lower reference temperature, has drifts when switched on and off. that is not so much the case with modified reference. but my assumption re the above is that your meter is always on, as I said) 732a references et al do have small short term drifts, you can determine them with a josephson element (these guys told me), but the 732a is certainly very stable short term compared to a 3458a. my 732a e.g. has a drift of 0.2ppm per year.
Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.
Randy
On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2688@gmail.com
wrote:
I am doing multiple 100 measurements simply to characterize the stability
of the 3458A and 732A units I just bought. After about 10 measurement sets
over 2 days I am seeing a variance of about .5 uV for the 10V output, or
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, acbern@gmx.de wrote:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am already
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal)
unless the amb. temperature is really very stable (half a degree already
adds about 0.25ppm at 10v)
thanks
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during
measurements. After the SMPL symbol no longer blinks I hit MATH 2
get a MATH ERR symbol on the display. I tried it a couple of times
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide,
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
set
the MATH to "Statistics" (store high reading/low reading/ and mean
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and PLC
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
got
a variation of 0.159 uVolts using the same 40 reading method above.
would
use this to determine where your problem might exist. Just having
meter
input shorted will point you in the right direction. Meter, cables
732A.
Sorry for the long dissertation. Friends get mad at me for
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads about
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack.
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be
if it is a distance away, or you have to use a shipper like UPS
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have the
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power
and then power is restored the result will be a different 10
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found
plugging in the lead will generate a thermal difference because
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A and
total difference of 0.159 uV over 40 measurements using 100 PLC
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
Randy,
You have two possible choices. It can be configured with 4 x
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
I have included a link to an ~30-hour measurement run that I consistently do that should give you some idea of expected measurement drift over time. Virtually all of the measurement drift is due to the 3458A internal temperature differences around the 7V reference caused by ambient temperature change. The drift associated with the 732A is probably about 2-magnitudes less at this ambient temp drift. In my situation, the inferred ambient temperature is cycling with the home air-conditioning. Note the initial calibration temperatures as well as the ACal temperatures and where the 3458A is measuring exactly 10V relative to those temperatures. My primary 732A has been powered without loss for 4 years and >5 years before that. The 3458A is Agilent (Loveland) Cal'ed yearly and is powered 24-7-365 except for the occasional mains power loss. The graphical measurements is using a homegrown Agilent VEE program. It is very helpful if not essential to get an HPIB interface setup so you can do long term graphical analysis.
The other link was done in winter time when the lab a few degrees cooler.
https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0
https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%203458A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0
I hope these Dropbox links work internationally - I'm new to using this sharing method.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Wednesday, August 27, 2014 8:42 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
well, what you do is to measure the stability of the 3458a. 1ppm drift though sounds much as overall short term average variation if your temperature is stable and your 3458a is always on. probably the temp is not stable, and that is what you see, amongst potentially some other smaller drifts like emf. before every cycle you should do an acal dcv, if you still see 10uv +/- drifts that then seems too much. (one thing I need to add is that one of my 3458a, not yet modified to have lower reference temperature, has drifts when switched on and off. that is not so much the case with modified reference. but my assumption re the above is that your meter is always on, as I said) 732a references et al do have small short term drifts, you can determine them with a josephson element (these guys told me), but the 732a is certainly very stable short term compared to a 3458a. my 732a e.g. has a drift of 0.2ppm per year.
> Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
> Von: "Randy Evans" <randyevans2688@gmail.com>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] 732A and Prologix received
>
> I forgot to mention that I reduced the number of measurements to 100 per
> set since I wasn't seeing much difference in the variance between 100 and
> 1000 measurements and the 1000 measurement per set takes too long.
>
> Randy
>
>
> On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > I am doing multiple 100 measurements simply to characterize the stability
> > of the 3458A and 732A units I just bought. After about 10 measurement sets
> > over 2 days I am seeing a variance of about .5 uV for the 10V output, or
> > 0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does
> > that sound reasonable/
> >
> > Randy
> >
> >
> > On Tue, Aug 26, 2014 at 5:41 PM, <acbern@gmx.de> wrote:
> >
> >> hi randy,
> >>
> >> just for curiosity, why doing 100 measurements at nplc 1000. is this to
> >> sample a changing value?
> >> when i am doing 10 measurements from a stable signal at nplc 100 (only
> >> there many subsequent measuremnts with statistics make sense) I am already
> >> getting a stanard deviation below 0.1ppm.
> >> in a 30 minute test cycle, i would also be concerned about drifts (acal)
> >> unless the amb. temperature is really very stable (half a degree already
> >> adds about 0.25ppm at 10v)
> >>
> >> thanks
> >>
> >>
> >>
> >>
> >> > Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
> >> > Von: "Randy Evans" <randyevans2688@gmail.com>
> >> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> > Betreff: Re: [volt-nuts] 732A and Prologix received
> >> >
> >> > Is there any way to tell when the function key routine is complete? In
> >> the
> >> > case of taking multiple readings using the DEFKEY and MATH function, I
> >> > don't see any indication when the routine is complete. In one
> >> particular
> >> > case, I am taking a 100 readings with NLPC set for 1000 so its a long
> >> while
> >> > before it's complete, but i have to guess when it's done.
> >> >
> >> > Thanks,
> >> >
> >> > Randy
> >> >
> >> >
> >> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> >> > wrote:
> >> >
> >> > > Bill,
> >> > >
> >> > > I am trying to figure out the MATH function without much success. I
> >> input
> >> > > the sequence you said (I looked up the instructions to understand
> >> what you
> >> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
> >> 4;TRIG; and
> >> > > it shows up on the display when I input BLUE F1. I hit ENTER and it
> >> takes
> >> > > the 40 measurements and the MATH symbol shows on the display during
> >> the
> >> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2
> >> and I
> >> > > get a MATH ERR symbol on the display. I tried it a couple of times
> >> and the
> >> > > same result so I am doing something wrong. Is there a better source
> >> for
> >> > > explaining how to do front panel masurements than the User Guide,
> >> which
> >> > > seems oriented at programming automatic rather than manual
> >> measurements.
> >> > >
> >> > > Randy
> >> > >
> >> > >
> >> > >
> >> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net>
> >> wrote:
> >> > >
> >> > >> Randy:
> >> > >>
> >> > >> The MATH function is accessible from the keypad. I don't have
> >> an IEEE
> >> > >> interface right now that works. You can also program the numeric
> >> keypad
> >> > >> keys to have preprogrammed functions. DEFKEY
> >> > >>
> >> > >> I have made my own "low thermal" measurement leads from Pomona
> >> #4892
> >> > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy
> >> at the
> >> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
> >> plans
> >> > >> to do custom cables with 16 ga. bare copper wire that I will twist
> >> and
> >> > >> then
> >> > >> put a braided shield over it. I simply cannot find what I want so I
> >> will
> >> > >> build my own cable. I have done something like this before and it
> >> worked
> >> > >> fine. When I get a "round toit".
> >> > >>
> >> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
> >> have
> >> > >> used
> >> > >> in the past to verify my homemade "low thermal" leads as described
> >> above.
> >> > >> Frankly I cannot see any difference between using the 1756 cables
> >> and my
> >> > >> homemade cables once I give them a few minutes for the thermals to go
> >> > >> away.
> >> > >> As far as I can tell and measure the differences, if any, are below
> >> 0.1
> >> > >> ppm
> >> > >> at 10 volts.
> >> > >>
> >> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
> >> are all
> >> > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
> >> far
> >> > >> as
> >> > >> the instability of the readings it is hard to determine which is
> >> causing
> >> > >> the
> >> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> >> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
> >> does is
> >> > >> set
> >> > >> the MATH to "Statistics" (store high reading/low reading/ and mean
> >> of the
> >> > >> readings) in the registers, the number of readings to "40", the
> >> trigger to
> >> > >> "hold" (which keeps the meter from triggering until I press "ENTER"
> >> and
> >> > >> then
> >> > >> trigger the sequence of 40 readings when I push the "ENTER" button.
> >> You
> >> > >> can
> >> > >> do all of this manually from the keypads but since I use this
> >> sequence a
> >> > >> lot
> >> > >> I have preprogrammed it. This is after I set digits to "8" and PLC
> >> to
> >> > >> 100.
> >> > >> Once those 40 readings are finished then you can access the various
> >> MATH
> >> > >> statistic registers, using the menu, by entering "MATH" and then a 2
> >> for
> >> > >> low, a 4 for mean, and 13 for high. Of course you could do all of
> >> this
> >> > >> through the IEEE also. The 3458A has a very rich set of measurement
> >> > >> commands. I am still learning all of them. It depends upon what I
> >> am
> >> > >> trying to accomplish.
> >> > >>
> >> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> >> > >> resistive dividers from the 10 volt, I don't see how they could
> >> contribute
> >> > >> to the varying readings you are measuring. I think I would put a
> >> short on
> >> > >> the input of the 3458A and manually set the range to 1 volt and then
> >> > >> observe
> >> > >> the variations that way without the 732A involved. When I do this I
> >> see a
> >> > >> variation from low reading to high reading of 0.125 uVolts and then
> >> > >> another
> >> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the
> >> low
> >> > >> side
> >> > >> of the measurment terminals, GUARD connected doesn't seem to affect
> >> the
> >> > >> readings. So that is the base noise of the 3458A without the 732A,
> >> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
> >> output I
> >> > >> got
> >> > >> a variation of 0.159 uVolts using the same 40 reading method above.
> >> I
> >> > >> would
> >> > >> use this to determine where your problem might exist. Just having
> >> the
> >> > >> meter
> >> > >> input shorted will point you in the right direction. Meter, cables
> >> or
> >> > >> 732A.
> >> > >>
> >> > >> Sorry for the long dissertation. Friends get mad at me for
> >> being so
> >> > >> detailed sometimes.
> >> > >>
> >> > >> Bill
> >> > >>
> >> > >> ----- Original Message -----
> >> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> >> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> > >> Sent: Sunday, August 24, 2014 7:22 AM
> >> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >>
> >> > >>
> >> > >> > Bill,
> >> > >> >
> >> > >> > I have convinced myself that the problem I an seeing is due to
> >> thermals.
> >> > >> > If I move the cables (with gold-plated banana plugs) using a small
> >> > >> towel
> >> > >> > rather than letting my hand touch the plugs, it is much more
> >> stable.
> >> > >> If I
> >> > >> > then hold the banana plug with my hand after the reading has
> >> stabilized,
> >> > >> > the reading drifts rapidly upward. I am trying to check the
> >> stability
> >> > >> of
> >> > >> > the reading but I haven't figured out the MATH function yet. I
> >> assume
> >> > >> this
> >> > >> > is a programmed function using GPIB only?
> >> > >> >
> >> > >> > The stability I am seeing by manually recording readings (using
> >> NLPC of
> >> > >> 100
> >> > >> > and 1000) is much greater than what you are measuring on your
> >> system.
> >> > >> Not
> >> > >> > sure how to ascertain if it's the 3458A or the 732. The value of
> >> the
> >> > >> > readings are very different between the two - the 3458 reads about
> >> 50 uV
> >> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> >> Rather
> >> > >> large
> >> > >> > differences (this is after an ACAL). I need to find some better
> >> cables
> >> > >> to
> >> > >> > make sure the errors are not due to thermals again.
> >> > >> >
> >> > >> > Randy
> >> > >> >
> >> > >> >
> >> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
> >> wrote:
> >> > >> >
> >> > >> > > Randy:
> >> > >> > >
> >> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
> >> will fit
> >> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
> >> LC-R064R5C and
> >> > >> > > others
> >> > >> > > that are in this size and package. Order from one of the usual
> >> > >> electronics
> >> > >> > > distributors like Allied, Mouser, Digikey. This is a very common
> >> > >> battery
> >> > >> > > as
> >> > >> > > it is used in a lot of "EXIT" signs so they are lighted when the
> >> power
> >> > >> goes
> >> > >> > > out. I don't see how the 12V 7AH will fit as they are too
> >> large. I
> >> > >> guess
> >> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
> >> volt
> >> > >> 4
> >> > >> AH
> >> > >> > > but the terminals are in the wrong place so you will have to
> >> "nibble"
> >> > >> out
> >> > >> > > the aluminum plate that holds them in the 732A battery pack.
> >> You have
> >> > >> to
> >> > >> > > be
> >> > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> >> > >> connection
> >> > >> > > leads to deal with and connect correctly. I would stick with
> >> the 6V
> >> > >> 4AH.
> >> > >> > > New batteries will last around 12 to 14 hours before the "CAL"
> >> light
> >> > >> goes
> >> > >> > > out when AC power is not applied. So shipping to Cal Lab can be
> >> a
> >> > >> problem
> >> > >> > > if it is a distance away, or you have to use a shipper like UPS
> >> or
> >> > >> FEDEX
> >> > >> > > and
> >> > >> > > you ship the night before and then use their "Morning delivery"
> >> and
> >> > >> the
> >> > >> Cal
> >> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> >> course
> >> > >> you
> >> > >> > > could always strap another battery on the 732A and hook it up to
> >> the
> >> > >> "ext
> >> > >> > > power" plug to last longer. I have seen it done. The issue is
> >> to get
> >> > >> the
> >> > >> > > Cal Lab to charge the extra battery before they ship the 732A
> >> back to
> >> > >> you.
> >> > >> > >
> >> > >> > > When you remove and work on the battery pack always have the
> >> AC
> >> > >> power
> >> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> >> regulated
> >> > >> supply
> >> > >> > > is working.
> >> > >> > >
> >> > >> > > The "CAL" light is to indicate that power has not been lost
> >> to the
> >> > >> > > Reference Amp or other associated circuits. When the raw supply
> >> > >> (battery)
> >> > >> > > voltage drops below about 21 volts the "CAL" light will go out.
> >> Below
> >> > >> that
> >> > >> > > voltage the heater circuits will not work correctly and the 18.6
> >> volt
> >> > >> > > regulated supply will not regulate. The requirement is that the
> >> > >> Reference
> >> > >> > > Amp be kept "alive" at all times to maintain the output voltage
> >> that
> >> > >> was
> >> > >> > > measured at the time of the most recent Calibration or
> >> Certification.
> >> > >> When
> >> > >> > > the semiconductor junctions are unbiased and cool off when power
> >> is
> >> > >> lost,
> >> > >> > > and then power is restored the result will be a different 10
> >> volts
> >> > >> than
> >> > >> > > before the power failure. My experience is that after all of the
> >> > >> years
> >> > >> > > that
> >> > >> > > these units have been powered up, this won't happen and when
> >> power is
> >> > >> lost
> >> > >> > > and then restored, even months later, the 732A will come back to
> >> > >> almost
> >> > >> > > exactly the same 10 volts as when they lost power, usually with
> >> in 0.2
> >> > >> PPM
> >> > >> > > after 24 hours of "warm up".
> >> > >> > >
> >> > >> > > What type of hookup leads are you using when measuring the 1
> >> volt
> >> > >> > > output? What is the PLC set to? I always use 100 PLC to measure
> >> > >> this.
> >> > >> If
> >> > >> > > you don't have "low thermal" connection leads you can experience
> >> uV
> >> > >> changes
> >> > >> > > for a minute or more after plugging in the leads due to the
> >> "thermals"
> >> > >> > > generated because of the difference in temperature between the
> >> banana
> >> > >> jacks
> >> > >> > > on the 732A and the banana plugs of the leads. I have found
> >> that even
> >> > >> just
> >> > >> > > plugging in the lead will generate a thermal difference because
> >> of
> >> > >> > > difference of temps and some heating due to the physical act of
> >> just
> >> > >> > > inserting the plug because of friction between the jack and plug
> >> (my
> >> > >> theory
> >> > >> > > at any rate). You have to allow at least a minute or more before
> >> > >> being
> >> > >> > > able
> >> > >> > > to make a measurement after plugging in the leads. I just
> >> measured
> >> > >> the
> >> > >> > > variation of the 1 volt output of my 732A and using my 3458A and
> >> I
> >> > >> got a
> >> > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC
> >> on
> >> > >> the
> >> > >> 1
> >> > >> > > volt range of the 3458A. Using the MATH function and all of the
> >> data
> >> > >> you
> >> > >> > > can collect. That was after waiting for several minutes after
> >> > >> plugging
> >> > >> in
> >> > >> > > the leads.
> >> > >> > >
> >> > >> > > I hope all of this helps.
> >> > >> > >
> >> > >> > > Bill
> >> > >> > >
> >> > >> > >
> >> > >> > >
> >> > >> > > ----- Original Message -----
> >> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> >> > >> > > To: "Discussion of precise voltage measurement" <
> >> volt-nuts@febo.com>
> >> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> >> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >> > >
> >> > >> > >
> >> > >> > > > Todd,
> >> > >> > > >
> >> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> >> batteries
> >> > >> that
> >> > >> I
> >> > >> > > > keep topped off and they have very low current draw (~2 to 3
> >> mA at
> >> > >> 13.5
> >> > >> > > > VDC) when charged and at their float voltage, so I am pretty
> >> sure
> >> > >> they
> >> > >> > > are
> >> > >> > > > in good condition. I will look at getting those in the units
> >> after
> >> > >> I
> >> > >> > > > ascertain the condition of the 732.
> >> > >> > > >
> >> > >> > > > So now I have a what appears to be a functioning 3458A and a
> >> 732A
> >> > >> but
> >> > >> > > they
> >> > >> > > > slightly disagree. I am like the man with two watches that
> >> disagree
> >> > >> on
> >> > >> > > the
> >> > >> > > > time - which is correct? For the moment, i am only concerned
> >> with
> >> > >> > > > stability. The need for absolute accuracy will come later.
> >> > >> > > >
> >> > >> > > > Randy
> >> > >> > > >
> >> > >> > > >
> >> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
> >> tmicallef@gmail.com
> >> > >> >
> >> > >> > > wrote:
> >> > >> > > >
> >> > >> > > > > Randy,
> >> > >> > > > >
> >> > >> > > > > You have two possible choices. It can be configured with 4 x
> >> 6v
> >> > >> 4Ah
> >> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> >> owner
> >> > >> has
> >> > >> > > > > modified the battery pack already. A couple of mine needed a
> >> > >> nibbler
> >> > >> > > tool
> >> > >> > > > > to remove enough of the aluminum cover that fits over the
> >> tops of
> >> > >> the
> >> > >> > > > > batteries. The original cover will short out to the battery
> >> tabs
> >> > >> > > regardless
> >> > >> > > > > of the battery configuration if this is not done.
> >> > >> > > > >
> >> > >> > > > > You can find larger capacity batteries that will give you
> >> slightly
> >> > >> more
> >> > >> > > > > battery life. I lost a couple sets of mail-order batteries
> >> after a
> >> > >> few
> >> > >> > > > > extended outages. I would recommend going with locally bought
> >> > >> batteries
> >> > >> > > > > instead of the cheaper mail order. My local Batteries Plus
> >> will
> >> > >> > > typically
> >> > >> > > > > have some warranty if I remember correctly. Moving forward I
> >> will
> >> > >> only
> >> > >> > > use
> >> > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> >> > >> equalize
> >> > >> > > them
> >> > >> > > > > before putting them in the 732A. I think the cheap batteries
> >> did
> >> > >> not
> >> > >> > > > > discharge equally, and would not recover when power was
> >> applied.
> >> > >> > > > >
> >> > >> > > > > Inspect the back plane for damaged traces and look at the
> >> > >> capacitors. I
> >> > >> > > had
> >> > >> > > > > a few that looked questionable. So far, I have replaced all
> >> the
> >> > >> big
> >> > >> > > caps
> >> > >> > > on
> >> > >> > > > > the pre-regulator and regulator boards. My feeling is that
> >> once
> >> > >> these
> >> > >> > > go
> >> > >> > > > > online, they should run as long as possible between repairs.
> >> > >> > > > >
> >> > >> > > > > The battery charger circuit may need adjusting. I tweaked
> >> mine and
> >> > >> it
> >> > >> > > > > seemed to work fine.
> >> > >> > > > >
> >> > >> > > > > Todd
> >> > >> > > > >
> >> > >> > > > >
> >> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> >> > >> > > randyevans2688@gmail.com>
> >> > >> > > > > wrote:
> >> > >> > > > >
> >> > >> > > > > > I received my Fluke 732A today. Just powered it up but it
> >> needs
> >> > >> new
> >> > >> > > > > > batteries. Any suggestions for sources (I haven't opened
> >> up the
> >> > >> unit
> >> > >> > > > > yet -
> >> > >> > > > > > I want to make sure it works before doing that). Also
> >> received
> >> > >> the
> >> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
> >> ran
> >> > >> data
> >> > >> > > > > dumper
> >> > >> > > > > > program to get the CAL data from my 3458A. Should be a
> >> busy
> >> > >> weekend.
> >> > >> > > > > >
> >> > >> > > > > > Randy
> >> > >> > > > > > _______________________________________________
> >> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > > > > To unsubscribe, go to
> >> > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > > > > and follow the instructions there.
> >> > >> > > > > >
> >> > >> > > > > _______________________________________________
> >> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > > > To unsubscribe, go to
> >> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > > > and follow the instructions there.
> >> > >> > > > >
> >> > >> > > > _______________________________________________
> >> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > > To unsubscribe, go to
> >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > > and follow the instructions there.
> >> > >> > >
> >> > >> > > _______________________________________________
> >> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > > To unsubscribe, go to
> >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > > and follow the instructions there.
> >> > >> > >
> >> > >> > _______________________________________________
> >> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> > To unsubscribe, go to
> >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> > and follow the instructions there.
> >> > >>
> >> > >> _______________________________________________
> >> > >> volt-nuts mailing list -- volt-nuts@febo.com
> >> > >> To unsubscribe, go to
> >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > >> and follow the instructions there.
> >> > >>
> >> > >
> >> > >
> >> > _______________________________________________
> >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > and follow the instructions there.
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >>
> >
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
RE
Randy Evans
Wed, Aug 27, 2014 3:37 PM
The HP3458A and the Fluke 732A are on continuously and I do an ACAL at
least every few hours, or when the room temperature changes by more than 1
degree C. The total range of measurements is 10uV so the drift is +/-5uV,
or 0.5 ppm. The room temperature is not particularly stable and varies
over a 3C range. The internal temp of the 3458 varies from 38.1 to 40.3
degrees C over the set of measurements. The 732A thermistor resistance
measures from 3.6677 Kohms to 3.6686 Kohms. I am using copper wires
between the 3458A and the 752A to minimize thermals. At the moment I have
no way to tell which unit is drifting the most.
Randy
On Wed, Aug 27, 2014 at 6:41 AM, acbern@gmx.de wrote:
well, what you do is to measure the stability of the 3458a. 1ppm drift
though sounds much as overall short term average variation if your
temperature is stable and your 3458a is always on. probably the temp is not
stable, and that is what you see, amongst potentially some other smaller
drifts like emf. before every cycle you should do an acal dcv, if you still
see 10uv +/- drifts that then seems too much. (one thing I need to add is
that one of my 3458a, not yet modified to have lower reference temperature,
has drifts when switched on and off. that is not so much the case with
modified reference. but my assumption re the above is that your meter is
always on, as I said)
732a references et al do have small short term drifts, you can determine
them with a josephson element (these guys told me), but the 732a is
certainly very stable short term compared to a 3458a. my 732a e.g. has a
drift of 0.2ppm per year.
Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.
Randy
On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2688@gmail.com
wrote:
I am doing multiple 100 measurements simply to characterize the
of the 3458A and 732A units I just bought. After about 10 measurement
over 2 days I am seeing a variance of about .5 uV for the 10V output,
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm.
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, acbern@gmx.de wrote:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts
unless the amb. temperature is really very stable (half a degree
adds about 0.25ppm at 10v)
thanks
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is
case of taking multiple readings using the DEFKEY and MATH
don't see any indication when the routine is complete. In one
case, I am taking a 100 readings with NLPC set for 1000 so its a
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <
Bill,
I am trying to figure out the MATH function without much
the sequence you said (I looked up the instructions to understand
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and
the 40 measurements and the MATH symbol shows on the display
measurements. After the SMPL symbol no longer blinks I hit MATH 2
get a MATH ERR symbol on the display. I tried it a couple of
same result so I am doing something wrong. Is there a better
explaining how to do front panel masurements than the User Guide,
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from
banana plugs and Belden #9272 wire. Why 9272, because it was
time. It is tin plated copper, shielded twisted pair 20 ga. I
to do custom cables with 16 ga. bare copper wire that I will
then
put a braided shield over it. I simply cannot find what I want
build my own cable. I have done something like this before and
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that
used
in the past to verify my homemade "low thermal" leads as
Frankly I cannot see any difference between using the 1756 cables
homemade cables once I give them a few minutes for the thermals
away.
As far as I can tell and measure the differences, if any, are
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the
adjustable you may be seeing a misadjusted 1 volt from the
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
set
the MATH to "Statistics" (store high reading/low reading/ and
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press
then
trigger the sequence of 40 readings when I push the "ENTER"
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and
Once those 40 readings are finished then you can access the
statistic registers, using the menu, by entering "MATH" and then
low, a 4 for mean, and 13 for high. Of course you could do all
through the IEEE also. The 3458A has a very rich set of
commands. I am still learning all of them. It depends upon
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and
observe
the variations that way without the 732A involved. When I do
variation from low reading to high reading of 0.125 uVolts and
another
40 I get 0.155 uVolts. This is without the GUARD connected to
side
of the measurment terminals, GUARD connected doesn't seem to
readings. So that is the base noise of the 3458A without the
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
got
a variation of 0.159 uVolts using the same 40 reading method
would
use this to determine where your problem might exist. Just
meter
input shorted will point you in the right direction. Meter,
732A.
Sorry for the long dissertation. Friends get mad at me for
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a
rather than letting my hand touch the plugs, it is much more
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value
readings are very different between the two - the 3458 reads
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
others
that are in this size and package. Order from one of the
distributors like Allied, Mouser, Digikey. This is a very
as
it is used in a lot of "EXIT" signs so they are lighted when
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack.
be
careful if you use the 12v 5AH as you will have 4 extra
leads to deal with and connect correctly. I would stick with
New batteries will last around 12 to 14 hours before the
out when AC power is not applied. So shipping to Cal Lab
if it is a distance away, or you have to use a shipper like
and
you ship the night before and then use their "Morning
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it
power" plug to last longer. I have seen it done. The issue
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been
Reference Amp or other associated circuits. When the raw
voltage drops below about 21 volts the "CAL" light will go
voltage the heater circuits will not work correctly and the
regulated supply will not regulate. The requirement is that
Amp be kept "alive" at all times to maintain the output
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when
and then power is restored the result will be a different 10
before the power failure. My experience is that after all
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come
exactly the same 10 volts as when they lost power, usually
after 24 hours of "warm up".
What type of hookup leads are you using when measuring
output? What is the PLC set to? I always use 100 PLC to
you don't have "low thermal" connection leads you can
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between
on the 732A and the banana plugs of the leads. I have found
plugging in the lead will generate a thermal difference
difference of temps and some heating due to the physical act
inserting the plug because of friction between the jack and
at any rate). You have to allow at least a minute or more
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A
total difference of 0.159 uV over 40 measurements using 100
volt range of the 3458A. Using the MATH function and all of
can collect. That was after waiting for several minutes
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to
VDC) when charged and at their float voltage, so I am
in good condition. I will look at getting those in the
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
Randy,
You have two possible choices. It can be configured with
batteries or 2 x 12v 7Ah batteries. Hopefully the
modified the battery pack already. A couple of mine
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order
extended outages. I would recommend going with locally
instead of the cheaper mail order. My local Batteries
have some warranty if I remember correctly. Moving
2 12v batteries and pre-charge them on a battery charger
before putting them in the 732A. I think the cheap
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced
the pre-regulator and regulator boards. My feeling is
online, they should run as long as possible between
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up
batteries. Any suggestions for sources (I haven't
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims'
program to get the CAL data from my 3458A. Should be a
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
The HP3458A and the Fluke 732A are on continuously and I do an ACAL at
least every few hours, or when the room temperature changes by more than 1
degree C. The total range of measurements is 10uV so the drift is +/-5uV,
or 0.5 ppm. The room temperature is not particularly stable and varies
over a 3C range. The internal temp of the 3458 varies from 38.1 to 40.3
degrees C over the set of measurements. The 732A thermistor resistance
measures from 3.6677 Kohms to 3.6686 Kohms. I am using copper wires
between the 3458A and the 752A to minimize thermals. At the moment I have
no way to tell which unit is drifting the most.
Randy
On Wed, Aug 27, 2014 at 6:41 AM, <acbern@gmx.de> wrote:
> well, what you do is to measure the stability of the 3458a. 1ppm drift
> though sounds much as overall short term average variation if your
> temperature is stable and your 3458a is always on. probably the temp is not
> stable, and that is what you see, amongst potentially some other smaller
> drifts like emf. before every cycle you should do an acal dcv, if you still
> see 10uv +/- drifts that then seems too much. (one thing I need to add is
> that one of my 3458a, not yet modified to have lower reference temperature,
> has drifts when switched on and off. that is not so much the case with
> modified reference. but my assumption re the above is that your meter is
> always on, as I said)
> 732a references et al do have small short term drifts, you can determine
> them with a josephson element (these guys told me), but the 732a is
> certainly very stable short term compared to a 3458a. my 732a e.g. has a
> drift of 0.2ppm per year.
>
>
> > Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
> > Von: "Randy Evans" <randyevans2688@gmail.com>
> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Betreff: Re: [volt-nuts] 732A and Prologix received
> >
> > I forgot to mention that I reduced the number of measurements to 100 per
> > set since I wasn't seeing much difference in the variance between 100 and
> > 1000 measurements and the 1000 measurement per set takes too long.
> >
> > Randy
> >
> >
> > On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > I am doing multiple 100 measurements simply to characterize the
> stability
> > > of the 3458A and 732A units I just bought. After about 10 measurement
> sets
> > > over 2 days I am seeing a variance of about .5 uV for the 10V output,
> or
> > > 0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm.
> Does
> > > that sound reasonable/
> > >
> > > Randy
> > >
> > >
> > > On Tue, Aug 26, 2014 at 5:41 PM, <acbern@gmx.de> wrote:
> > >
> > >> hi randy,
> > >>
> > >> just for curiosity, why doing 100 measurements at nplc 1000. is this
> to
> > >> sample a changing value?
> > >> when i am doing 10 measurements from a stable signal at nplc 100 (only
> > >> there many subsequent measuremnts with statistics make sense) I am
> already
> > >> getting a stanard deviation below 0.1ppm.
> > >> in a 30 minute test cycle, i would also be concerned about drifts
> (acal)
> > >> unless the amb. temperature is really very stable (half a degree
> already
> > >> adds about 0.25ppm at 10v)
> > >>
> > >> thanks
> > >>
> > >>
> > >>
> > >>
> > >> > Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
> > >> > Von: "Randy Evans" <randyevans2688@gmail.com>
> > >> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com
> >
> > >> > Betreff: Re: [volt-nuts] 732A and Prologix received
> > >> >
> > >> > Is there any way to tell when the function key routine is
> complete? In
> > >> the
> > >> > case of taking multiple readings using the DEFKEY and MATH
> function, I
> > >> > don't see any indication when the routine is complete. In one
> > >> particular
> > >> > case, I am taking a 100 readings with NLPC set for 1000 so its a
> long
> > >> while
> > >> > before it's complete, but i have to guess when it's done.
> > >> >
> > >> > Thanks,
> > >> >
> > >> > Randy
> > >> >
> > >> >
> > >> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <
> randyevans2688@gmail.com>
> > >> > wrote:
> > >> >
> > >> > > Bill,
> > >> > >
> > >> > > I am trying to figure out the MATH function without much
> success. I
> > >> input
> > >> > > the sequence you said (I looked up the instructions to understand
> > >> what you
> > >> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
> > >> 4;TRIG; and
> > >> > > it shows up on the display when I input BLUE F1. I hit ENTER and
> it
> > >> takes
> > >> > > the 40 measurements and the MATH symbol shows on the display
> during
> > >> the
> > >> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2
> > >> and I
> > >> > > get a MATH ERR symbol on the display. I tried it a couple of
> times
> > >> and the
> > >> > > same result so I am doing something wrong. Is there a better
> source
> > >> for
> > >> > > explaining how to do front panel masurements than the User Guide,
> > >> which
> > >> > > seems oriented at programming automatic rather than manual
> > >> measurements.
> > >> > >
> > >> > > Randy
> > >> > >
> > >> > >
> > >> > >
> > >> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net>
> > >> wrote:
> > >> > >
> > >> > >> Randy:
> > >> > >>
> > >> > >> The MATH function is accessible from the keypad. I don't
> have
> > >> an IEEE
> > >> > >> interface right now that works. You can also program the numeric
> > >> keypad
> > >> > >> keys to have preprogrammed functions. DEFKEY
> > >> > >>
> > >> > >> I have made my own "low thermal" measurement leads from
> Pomona
> > >> #4892
> > >> > >> banana plugs and Belden #9272 wire. Why 9272, because it was
> handy
> > >> at the
> > >> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I
> have
> > >> plans
> > >> > >> to do custom cables with 16 ga. bare copper wire that I will
> twist
> > >> and
> > >> > >> then
> > >> > >> put a braided shield over it. I simply cannot find what I want
> so I
> > >> will
> > >> > >> build my own cable. I have done something like this before and
> it
> > >> worked
> > >> > >> fine. When I get a "round toit".
> > >> > >>
> > >> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that
> I
> > >> have
> > >> > >> used
> > >> > >> in the past to verify my homemade "low thermal" leads as
> described
> > >> above.
> > >> > >> Frankly I cannot see any difference between using the 1756 cables
> > >> and my
> > >> > >> homemade cables once I give them a few minutes for the thermals
> to go
> > >> > >> away.
> > >> > >> As far as I can tell and measure the differences, if any, are
> below
> > >> 0.1
> > >> > >> ppm
> > >> > >> at 10 volts.
> > >> > >>
> > >> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the
> 732A
> > >> are all
> > >> > >> adjustable you may be seeing a misadjusted 1 volt from the
> 732A. As
> > >> far
> > >> > >> as
> > >> > >> the instability of the readings it is hard to determine which is
> > >> causing
> > >> > >> the
> > >> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with
> the
> > >> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
> > >> does is
> > >> > >> set
> > >> > >> the MATH to "Statistics" (store high reading/low reading/ and
> mean
> > >> of the
> > >> > >> readings) in the registers, the number of readings to "40", the
> > >> trigger to
> > >> > >> "hold" (which keeps the meter from triggering until I press
> "ENTER"
> > >> and
> > >> > >> then
> > >> > >> trigger the sequence of 40 readings when I push the "ENTER"
> button.
> > >> You
> > >> > >> can
> > >> > >> do all of this manually from the keypads but since I use this
> > >> sequence a
> > >> > >> lot
> > >> > >> I have preprogrammed it. This is after I set digits to "8" and
> PLC
> > >> to
> > >> > >> 100.
> > >> > >> Once those 40 readings are finished then you can access the
> various
> > >> MATH
> > >> > >> statistic registers, using the menu, by entering "MATH" and then
> a 2
> > >> for
> > >> > >> low, a 4 for mean, and 13 for high. Of course you could do all
> of
> > >> this
> > >> > >> through the IEEE also. The 3458A has a very rich set of
> measurement
> > >> > >> commands. I am still learning all of them. It depends upon
> what I
> > >> am
> > >> > >> trying to accomplish.
> > >> > >>
> > >> > >> Since the 1.018 and 1.0 volt outputs are passive and derived
> from
> > >> > >> resistive dividers from the 10 volt, I don't see how they could
> > >> contribute
> > >> > >> to the varying readings you are measuring. I think I would put a
> > >> short on
> > >> > >> the input of the 3458A and manually set the range to 1 volt and
> then
> > >> > >> observe
> > >> > >> the variations that way without the 732A involved. When I do
> this I
> > >> see a
> > >> > >> variation from low reading to high reading of 0.125 uVolts and
> then
> > >> > >> another
> > >> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to
> the
> > >> low
> > >> > >> side
> > >> > >> of the measurment terminals, GUARD connected doesn't seem to
> affect
> > >> the
> > >> > >> readings. So that is the base noise of the 3458A without the
> 732A,
> > >> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
> > >> output I
> > >> > >> got
> > >> > >> a variation of 0.159 uVolts using the same 40 reading method
> above.
> > >> I
> > >> > >> would
> > >> > >> use this to determine where your problem might exist. Just
> having
> > >> the
> > >> > >> meter
> > >> > >> input shorted will point you in the right direction. Meter,
> cables
> > >> or
> > >> > >> 732A.
> > >> > >>
> > >> > >> Sorry for the long dissertation. Friends get mad at me for
> > >> being so
> > >> > >> detailed sometimes.
> > >> > >>
> > >> > >> Bill
> > >> > >>
> > >> > >> ----- Original Message -----
> > >> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > >> To: "Discussion of precise voltage measurement" <
> volt-nuts@febo.com>
> > >> > >> Sent: Sunday, August 24, 2014 7:22 AM
> > >> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >>
> > >> > >>
> > >> > >> > Bill,
> > >> > >> >
> > >> > >> > I have convinced myself that the problem I an seeing is due to
> > >> thermals.
> > >> > >> > If I move the cables (with gold-plated banana plugs) using a
> small
> > >> > >> towel
> > >> > >> > rather than letting my hand touch the plugs, it is much more
> > >> stable.
> > >> > >> If I
> > >> > >> > then hold the banana plug with my hand after the reading has
> > >> stabilized,
> > >> > >> > the reading drifts rapidly upward. I am trying to check the
> > >> stability
> > >> > >> of
> > >> > >> > the reading but I haven't figured out the MATH function yet. I
> > >> assume
> > >> > >> this
> > >> > >> > is a programmed function using GPIB only?
> > >> > >> >
> > >> > >> > The stability I am seeing by manually recording readings (using
> > >> NLPC of
> > >> > >> 100
> > >> > >> > and 1000) is much greater than what you are measuring on your
> > >> system.
> > >> > >> Not
> > >> > >> > sure how to ascertain if it's the 3458A or the 732. The value
> of
> > >> the
> > >> > >> > readings are very different between the two - the 3458 reads
> about
> > >> 50 uV
> > >> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> > >> Rather
> > >> > >> large
> > >> > >> > differences (this is after an ACAL). I need to find some
> better
> > >> cables
> > >> > >> to
> > >> > >> > make sure the errors are not due to thermals again.
> > >> > >> >
> > >> > >> > Randy
> > >> > >> >
> > >> > >> >
> > >> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <
> wpgold3637@att.net>
> > >> wrote:
> > >> > >> >
> > >> > >> > > Randy:
> > >> > >> > >
> > >> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
> > >> will fit
> > >> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
> > >> LC-R064R5C and
> > >> > >> > > others
> > >> > >> > > that are in this size and package. Order from one of the
> usual
> > >> > >> electronics
> > >> > >> > > distributors like Allied, Mouser, Digikey. This is a very
> common
> > >> > >> battery
> > >> > >> > > as
> > >> > >> > > it is used in a lot of "EXIT" signs so they are lighted when
> the
> > >> power
> > >> > >> goes
> > >> > >> > > out. I don't see how the 12V 7AH will fit as they are too
> > >> large. I
> > >> > >> guess
> > >> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2
> x 6
> > >> volt
> > >> > >> 4
> > >> > >> AH
> > >> > >> > > but the terminals are in the wrong place so you will have to
> > >> "nibble"
> > >> > >> out
> > >> > >> > > the aluminum plate that holds them in the 732A battery pack.
> > >> You have
> > >> > >> to
> > >> > >> > > be
> > >> > >> > > careful if you use the 12v 5AH as you will have 4 extra
> battery
> > >> > >> connection
> > >> > >> > > leads to deal with and connect correctly. I would stick with
> > >> the 6V
> > >> > >> 4AH.
> > >> > >> > > New batteries will last around 12 to 14 hours before the
> "CAL"
> > >> light
> > >> > >> goes
> > >> > >> > > out when AC power is not applied. So shipping to Cal Lab
> can be
> > >> a
> > >> > >> problem
> > >> > >> > > if it is a distance away, or you have to use a shipper like
> UPS
> > >> or
> > >> > >> FEDEX
> > >> > >> > > and
> > >> > >> > > you ship the night before and then use their "Morning
> delivery"
> > >> and
> > >> > >> the
> > >> > >> Cal
> > >> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> > >> course
> > >> > >> you
> > >> > >> > > could always strap another battery on the 732A and hook it
> up to
> > >> the
> > >> > >> "ext
> > >> > >> > > power" plug to last longer. I have seen it done. The issue
> is
> > >> to get
> > >> > >> the
> > >> > >> > > Cal Lab to charge the extra battery before they ship the 732A
> > >> back to
> > >> > >> you.
> > >> > >> > >
> > >> > >> > > When you remove and work on the battery pack always have
> the
> > >> AC
> > >> > >> power
> > >> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> > >> regulated
> > >> > >> supply
> > >> > >> > > is working.
> > >> > >> > >
> > >> > >> > > The "CAL" light is to indicate that power has not been
> lost
> > >> to the
> > >> > >> > > Reference Amp or other associated circuits. When the raw
> supply
> > >> > >> (battery)
> > >> > >> > > voltage drops below about 21 volts the "CAL" light will go
> out.
> > >> Below
> > >> > >> that
> > >> > >> > > voltage the heater circuits will not work correctly and the
> 18.6
> > >> volt
> > >> > >> > > regulated supply will not regulate. The requirement is that
> the
> > >> > >> Reference
> > >> > >> > > Amp be kept "alive" at all times to maintain the output
> voltage
> > >> that
> > >> > >> was
> > >> > >> > > measured at the time of the most recent Calibration or
> > >> Certification.
> > >> > >> When
> > >> > >> > > the semiconductor junctions are unbiased and cool off when
> power
> > >> is
> > >> > >> lost,
> > >> > >> > > and then power is restored the result will be a different 10
> > >> volts
> > >> > >> than
> > >> > >> > > before the power failure. My experience is that after all
> of the
> > >> > >> years
> > >> > >> > > that
> > >> > >> > > these units have been powered up, this won't happen and when
> > >> power is
> > >> > >> lost
> > >> > >> > > and then restored, even months later, the 732A will come
> back to
> > >> > >> almost
> > >> > >> > > exactly the same 10 volts as when they lost power, usually
> with
> > >> in 0.2
> > >> > >> PPM
> > >> > >> > > after 24 hours of "warm up".
> > >> > >> > >
> > >> > >> > > What type of hookup leads are you using when measuring
> the 1
> > >> volt
> > >> > >> > > output? What is the PLC set to? I always use 100 PLC to
> measure
> > >> > >> this.
> > >> > >> If
> > >> > >> > > you don't have "low thermal" connection leads you can
> experience
> > >> uV
> > >> > >> changes
> > >> > >> > > for a minute or more after plugging in the leads due to the
> > >> "thermals"
> > >> > >> > > generated because of the difference in temperature between
> the
> > >> banana
> > >> > >> jacks
> > >> > >> > > on the 732A and the banana plugs of the leads. I have found
> > >> that even
> > >> > >> just
> > >> > >> > > plugging in the lead will generate a thermal difference
> because
> > >> of
> > >> > >> > > difference of temps and some heating due to the physical act
> of
> > >> just
> > >> > >> > > inserting the plug because of friction between the jack and
> plug
> > >> (my
> > >> > >> theory
> > >> > >> > > at any rate). You have to allow at least a minute or more
> before
> > >> > >> being
> > >> > >> > > able
> > >> > >> > > to make a measurement after plugging in the leads. I just
> > >> measured
> > >> > >> the
> > >> > >> > > variation of the 1 volt output of my 732A and using my 3458A
> and
> > >> I
> > >> > >> got a
> > >> > >> > > total difference of 0.159 uV over 40 measurements using 100
> PLC
> > >> on
> > >> > >> the
> > >> > >> 1
> > >> > >> > > volt range of the 3458A. Using the MATH function and all of
> the
> > >> data
> > >> > >> you
> > >> > >> > > can collect. That was after waiting for several minutes
> after
> > >> > >> plugging
> > >> > >> in
> > >> > >> > > the leads.
> > >> > >> > >
> > >> > >> > > I hope all of this helps.
> > >> > >> > >
> > >> > >> > > Bill
> > >> > >> > >
> > >> > >> > >
> > >> > >> > >
> > >> > >> > > ----- Original Message -----
> > >> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > >> > > To: "Discussion of precise voltage measurement" <
> > >> volt-nuts@febo.com>
> > >> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> > >> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >> > >
> > >> > >> > >
> > >> > >> > > > Todd,
> > >> > >> > > >
> > >> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> > >> batteries
> > >> > >> that
> > >> > >> I
> > >> > >> > > > keep topped off and they have very low current draw (~2 to
> 3
> > >> mA at
> > >> > >> 13.5
> > >> > >> > > > VDC) when charged and at their float voltage, so I am
> pretty
> > >> sure
> > >> > >> they
> > >> > >> > > are
> > >> > >> > > > in good condition. I will look at getting those in the
> units
> > >> after
> > >> > >> I
> > >> > >> > > > ascertain the condition of the 732.
> > >> > >> > > >
> > >> > >> > > > So now I have a what appears to be a functioning 3458A and
> a
> > >> 732A
> > >> > >> but
> > >> > >> > > they
> > >> > >> > > > slightly disagree. I am like the man with two watches that
> > >> disagree
> > >> > >> on
> > >> > >> > > the
> > >> > >> > > > time - which is correct? For the moment, i am only
> concerned
> > >> with
> > >> > >> > > > stability. The need for absolute accuracy will come later.
> > >> > >> > > >
> > >> > >> > > > Randy
> > >> > >> > > >
> > >> > >> > > >
> > >> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
> > >> tmicallef@gmail.com
> > >> > >> >
> > >> > >> > > wrote:
> > >> > >> > > >
> > >> > >> > > > > Randy,
> > >> > >> > > > >
> > >> > >> > > > > You have two possible choices. It can be configured with
> 4 x
> > >> 6v
> > >> > >> 4Ah
> > >> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the
> previous
> > >> owner
> > >> > >> has
> > >> > >> > > > > modified the battery pack already. A couple of mine
> needed a
> > >> > >> nibbler
> > >> > >> > > tool
> > >> > >> > > > > to remove enough of the aluminum cover that fits over the
> > >> tops of
> > >> > >> the
> > >> > >> > > > > batteries. The original cover will short out to the
> battery
> > >> tabs
> > >> > >> > > regardless
> > >> > >> > > > > of the battery configuration if this is not done.
> > >> > >> > > > >
> > >> > >> > > > > You can find larger capacity batteries that will give you
> > >> slightly
> > >> > >> more
> > >> > >> > > > > battery life. I lost a couple sets of mail-order
> batteries
> > >> after a
> > >> > >> few
> > >> > >> > > > > extended outages. I would recommend going with locally
> bought
> > >> > >> batteries
> > >> > >> > > > > instead of the cheaper mail order. My local Batteries
> Plus
> > >> will
> > >> > >> > > typically
> > >> > >> > > > > have some warranty if I remember correctly. Moving
> forward I
> > >> will
> > >> > >> only
> > >> > >> > > use
> > >> > >> > > > > 2 12v batteries and pre-charge them on a battery charger
> to
> > >> > >> equalize
> > >> > >> > > them
> > >> > >> > > > > before putting them in the 732A. I think the cheap
> batteries
> > >> did
> > >> > >> not
> > >> > >> > > > > discharge equally, and would not recover when power was
> > >> applied.
> > >> > >> > > > >
> > >> > >> > > > > Inspect the back plane for damaged traces and look at the
> > >> > >> capacitors. I
> > >> > >> > > had
> > >> > >> > > > > a few that looked questionable. So far, I have replaced
> all
> > >> the
> > >> > >> big
> > >> > >> > > caps
> > >> > >> > > on
> > >> > >> > > > > the pre-regulator and regulator boards. My feeling is
> that
> > >> once
> > >> > >> these
> > >> > >> > > go
> > >> > >> > > > > online, they should run as long as possible between
> repairs.
> > >> > >> > > > >
> > >> > >> > > > > The battery charger circuit may need adjusting. I tweaked
> > >> mine and
> > >> > >> it
> > >> > >> > > > > seemed to work fine.
> > >> > >> > > > >
> > >> > >> > > > > Todd
> > >> > >> > > > >
> > >> > >> > > > >
> > >> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > >> > >> > > randyevans2688@gmail.com>
> > >> > >> > > > > wrote:
> > >> > >> > > > >
> > >> > >> > > > > > I received my Fluke 732A today. Just powered it up
> but it
> > >> needs
> > >> > >> new
> > >> > >> > > > > > batteries. Any suggestions for sources (I haven't
> opened
> > >> up the
> > >> > >> unit
> > >> > >> > > > > yet -
> > >> > >> > > > > > I want to make sure it works before doing that). Also
> > >> received
> > >> > >> the
> > >> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims'
> CAL
> > >> ran
> > >> > >> data
> > >> > >> > > > > dumper
> > >> > >> > > > > > program to get the CAL data from my 3458A. Should be a
> > >> busy
> > >> > >> weekend.
> > >> > >> > > > > >
> > >> > >> > > > > > Randy
> > >> > >> > > > > > _______________________________________________
> > >> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > > > > To unsubscribe, go to
> > >> > >> > > > > >
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > > > > and follow the instructions there.
> > >> > >> > > > > >
> > >> > >> > > > > _______________________________________________
> > >> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > > > To unsubscribe, go to
> > >> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > > > and follow the instructions there.
> > >> > >> > > > >
> > >> > >> > > > _______________________________________________
> > >> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > > To unsubscribe, go to
> > >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > > and follow the instructions there.
> > >> > >> > >
> > >> > >> > > _______________________________________________
> > >> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > To unsubscribe, go to
> > >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > and follow the instructions there.
> > >> > >> > >
> > >> > >> > _______________________________________________
> > >> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > To unsubscribe, go to
> > >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > and follow the instructions there.
> > >> > >>
> > >> > >> _______________________________________________
> > >> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> To unsubscribe, go to
> > >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> and follow the instructions there.
> > >> > >>
> > >> > >
> > >> > >
> > >> > _______________________________________________
> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > and follow the instructions there.
> > >> _______________________________________________
> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> and follow the instructions there.
> > >>
> > >
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
RE
Randy Evans
Wed, Aug 27, 2014 3:49 PM
Don,
If I am reading your charts correctly, it looks like your system is cycling
over an 7 uV range over several days in summer, similar to what I am
seeing. I don't have air conditioning (Northern California) so I do get
some pretty good temperature variations from day to night, but less inside
the house.
I need to figure out how to share files in Dropbox since I don't have my
own web site. Good idea.
thanks,
Randy
On Wed, Aug 27, 2014 at 7:40 AM, Don@True-Cal truecalservices@gmail.com
wrote:
I have included a link to an ~30-hour measurement run that I consistently
do that should give you some idea of expected measurement drift over time.
Virtually all of the measurement drift is due to the 3458A internal
temperature differences around the 7V reference caused by ambient
temperature change. The drift associated with the 732A is probably about
2-magnitudes less at this ambient temp drift. In my situation, the inferred
ambient temperature is cycling with the home air-conditioning. Note the
initial calibration temperatures as well as the ACal temperatures and where
the 3458A is measuring exactly 10V relative to those temperatures. My
primary 732A has been powered without loss for 4 years and >5 years before
that. The 3458A is Agilent (Loveland) Cal'ed yearly and is powered 24-7-365
except for the occasional mains power loss. The graphical measurements is
using a homegrown Agilent VEE program. It is very helpful if not essential
to get an HPIB interface setup so you can do long term graphical analysis.
The other link was done in winter time when the lab a few degrees cooler.
https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0
https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%203458A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0
I hope these Dropbox links work internationally - I'm new to using this
sharing method.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of
acbern@gmx.de
Sent: Wednesday, August 27, 2014 8:42 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received
well, what you do is to measure the stability of the 3458a. 1ppm drift
though sounds much as overall short term average variation if your
temperature is stable and your 3458a is always on. probably the temp is not
stable, and that is what you see, amongst potentially some other smaller
drifts like emf. before every cycle you should do an acal dcv, if you still
see 10uv +/- drifts that then seems too much. (one thing I need to add is
that one of my 3458a, not yet modified to have lower reference temperature,
has drifts when switched on and off. that is not so much the case with
modified reference. but my assumption re the above is that your meter is
always on, as I said) 732a references et al do have small short term
drifts, you can determine them with a josephson element (these guys told
me), but the 732a is certainly very stable short term compared to a 3458a.
my 732a e.g. has a drift of 0.2ppm per year.
Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
Von: "Randy Evans" randyevans2688@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.
Randy
On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2688@gmail.com
wrote:
I am doing multiple 100 measurements simply to characterize the
of the 3458A and 732A units I just bought. After about 10 measurement
over 2 days I am seeing a variance of about .5 uV for the 10V output,
0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm.
that sound reasonable/
Randy
On Tue, Aug 26, 2014 at 5:41 PM, acbern@gmx.de wrote:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts
unless the amb. temperature is really very stable (half a degree
adds about 0.25ppm at 10v)
thanks
Betreff: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is
case of taking multiple readings using the DEFKEY and MATH
don't see any indication when the routine is complete. In one
case, I am taking a 100 readings with NLPC set for 1000 so its a
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <
Bill,
I am trying to figure out the MATH function without much
the sequence you said (I looked up the instructions to understand
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
it shows up on the display when I input BLUE F1. I hit ENTER and
the 40 measurements and the MATH symbol shows on the display
measurements. After the SMPL symbol no longer blinks I hit MATH 2
get a MATH ERR symbol on the display. I tried it a couple of
same result so I am doing something wrong. Is there a better
explaining how to do front panel masurements than the User Guide,
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from
banana plugs and Belden #9272 wire. Why 9272, because it was
time. It is tin plated copper, shielded twisted pair 20 ga. I
to do custom cables with 16 ga. bare copper wire that I will
then
put a braided shield over it. I simply cannot find what I want
build my own cable. I have done something like this before and
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that
used
in the past to verify my homemade "low thermal" leads as
Frankly I cannot see any difference between using the 1756 cables
homemade cables once I give them a few minutes for the thermals
away.
As far as I can tell and measure the differences, if any, are
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the
adjustable you may be seeing a misadjusted 1 volt from the
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
set
the MATH to "Statistics" (store high reading/low reading/ and
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press
then
trigger the sequence of 40 readings when I push the "ENTER"
can
do all of this manually from the keypads but since I use this
lot
I have preprogrammed it. This is after I set digits to "8" and
Once those 40 readings are finished then you can access the
statistic registers, using the menu, by entering "MATH" and then
low, a 4 for mean, and 13 for high. Of course you could do all
through the IEEE also. The 3458A has a very rich set of
commands. I am still learning all of them. It depends upon
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and
observe
the variations that way without the 732A involved. When I do
variation from low reading to high reading of 0.125 uVolts and
another
40 I get 0.155 uVolts. This is without the GUARD connected to
side
of the measurment terminals, GUARD connected doesn't seem to
readings. So that is the base noise of the 3458A without the
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
got
a variation of 0.159 uVolts using the same 40 reading method
would
use this to determine where your problem might exist. Just
meter
input shorted will point you in the right direction. Meter,
732A.
Sorry for the long dissertation. Friends get mad at me for
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a
rather than letting my hand touch the plugs, it is much more
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value
readings are very different between the two - the 3458 reads
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
others
that are in this size and package. Order from one of the
distributors like Allied, Mouser, Digikey. This is a very
as
it is used in a lot of "EXIT" signs so they are lighted when
out. I don't see how the 12V 7AH will fit as they are too
you could use a 12V 5AH (PS1250) as it is the same size as 2
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack.
be
careful if you use the 12v 5AH as you will have 4 extra
leads to deal with and connect correctly. I would stick with
New batteries will last around 12 to 14 hours before the
out when AC power is not applied. So shipping to Cal Lab
if it is a distance away, or you have to use a shipper like
and
you ship the night before and then use their "Morning
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it
power" plug to last longer. I have seen it done. The issue
Cal Lab to charge the extra battery before they ship the 732A
When you remove and work on the battery pack always have
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been
Reference Amp or other associated circuits. When the raw
voltage drops below about 21 volts the "CAL" light will go
voltage the heater circuits will not work correctly and the
regulated supply will not regulate. The requirement is that
Amp be kept "alive" at all times to maintain the output
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when
and then power is restored the result will be a different 10
before the power failure. My experience is that after all
that
these units have been powered up, this won't happen and when
and then restored, even months later, the 732A will come
exactly the same 10 volts as when they lost power, usually
after 24 hours of "warm up".
What type of hookup leads are you using when measuring
output? What is the PLC set to? I always use 100 PLC to
you don't have "low thermal" connection leads you can
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between
on the 732A and the banana plugs of the leads. I have found
plugging in the lead will generate a thermal difference
difference of temps and some heating due to the physical act
inserting the plug because of friction between the jack and
at any rate). You have to allow at least a minute or more
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A
total difference of 0.159 uV over 40 measurements using 100
volt range of the 3458A. Using the MATH function and all of
can collect. That was after waiting for several minutes
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" <
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to
VDC) when charged and at their float voltage, so I am
in good condition. I will look at getting those in the
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
Randy,
You have two possible choices. It can be configured with
batteries or 2 x 12v 7Ah batteries. Hopefully the
modified the battery pack already. A couple of mine
to remove enough of the aluminum cover that fits over the
batteries. The original cover will short out to the
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order
extended outages. I would recommend going with locally
instead of the cheaper mail order. My local Batteries
have some warranty if I remember correctly. Moving
2 12v batteries and pre-charge them on a battery charger
before putting them in the 732A. I think the cheap
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced
the pre-regulator and regulator boards. My feeling is
online, they should run as long as possible between
The battery charger circuit may need adjusting. I tweaked
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up
batteries. Any suggestions for sources (I haven't
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims'
program to get the CAL data from my 3458A. Should be a
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Don,
If I am reading your charts correctly, it looks like your system is cycling
over an 7 uV range over several days in summer, similar to what I am
seeing. I don't have air conditioning (Northern California) so I do get
some pretty good temperature variations from day to night, but less inside
the house.
I need to figure out how to share files in Dropbox since I don't have my
own web site. Good idea.
thanks,
Randy
On Wed, Aug 27, 2014 at 7:40 AM, Don@True-Cal <truecalservices@gmail.com>
wrote:
> I have included a link to an ~30-hour measurement run that I consistently
> do that should give you some idea of expected measurement drift over time.
> Virtually all of the measurement drift is due to the 3458A internal
> temperature differences around the 7V reference caused by ambient
> temperature change. The drift associated with the 732A is probably about
> 2-magnitudes less at this ambient temp drift. In my situation, the inferred
> ambient temperature is cycling with the home air-conditioning. Note the
> initial calibration temperatures as well as the ACal temperatures and where
> the 3458A is measuring exactly 10V relative to those temperatures. My
> primary 732A has been powered without loss for 4 years and >5 years before
> that. The 3458A is Agilent (Loveland) Cal'ed yearly and is powered 24-7-365
> except for the occasional mains power loss. The graphical measurements is
> using a homegrown Agilent VEE program. It is very helpful if not essential
> to get an HPIB interface setup so you can do long term graphical analysis.
>
> The other link was done in winter time when the lab a few degrees cooler.
>
>
> https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0
>
>
> https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%203458A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0
>
> I hope these Dropbox links work internationally - I'm new to using this
> sharing method.
>
> Don Johnson
>
> -----Original Message-----
> From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of
> acbern@gmx.de
> Sent: Wednesday, August 27, 2014 8:42 AM
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] 732A and Prologix received
>
> well, what you do is to measure the stability of the 3458a. 1ppm drift
> though sounds much as overall short term average variation if your
> temperature is stable and your 3458a is always on. probably the temp is not
> stable, and that is what you see, amongst potentially some other smaller
> drifts like emf. before every cycle you should do an acal dcv, if you still
> see 10uv +/- drifts that then seems too much. (one thing I need to add is
> that one of my 3458a, not yet modified to have lower reference temperature,
> has drifts when switched on and off. that is not so much the case with
> modified reference. but my assumption re the above is that your meter is
> always on, as I said) 732a references et al do have small short term
> drifts, you can determine them with a josephson element (these guys told
> me), but the 732a is certainly very stable short term compared to a 3458a.
> my 732a e.g. has a drift of 0.2ppm per year.
>
>
> > Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
> > Von: "Randy Evans" <randyevans2688@gmail.com>
> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > Betreff: Re: [volt-nuts] 732A and Prologix received
> >
> > I forgot to mention that I reduced the number of measurements to 100 per
> > set since I wasn't seeing much difference in the variance between 100 and
> > 1000 measurements and the 1000 measurement per set takes too long.
> >
> > Randy
> >
> >
> > On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > I am doing multiple 100 measurements simply to characterize the
> stability
> > > of the 3458A and 732A units I just bought. After about 10 measurement
> sets
> > > over 2 days I am seeing a variance of about .5 uV for the 10V output,
> or
> > > 0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm.
> Does
> > > that sound reasonable/
> > >
> > > Randy
> > >
> > >
> > > On Tue, Aug 26, 2014 at 5:41 PM, <acbern@gmx.de> wrote:
> > >
> > >> hi randy,
> > >>
> > >> just for curiosity, why doing 100 measurements at nplc 1000. is this
> to
> > >> sample a changing value?
> > >> when i am doing 10 measurements from a stable signal at nplc 100 (only
> > >> there many subsequent measuremnts with statistics make sense) I am
> already
> > >> getting a stanard deviation below 0.1ppm.
> > >> in a 30 minute test cycle, i would also be concerned about drifts
> (acal)
> > >> unless the amb. temperature is really very stable (half a degree
> already
> > >> adds about 0.25ppm at 10v)
> > >>
> > >> thanks
> > >>
> > >>
> > >>
> > >>
> > >> > Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
> > >> > Von: "Randy Evans" <randyevans2688@gmail.com>
> > >> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com
> >
> > >> > Betreff: Re: [volt-nuts] 732A and Prologix received
> > >> >
> > >> > Is there any way to tell when the function key routine is
> complete? In
> > >> the
> > >> > case of taking multiple readings using the DEFKEY and MATH
> function, I
> > >> > don't see any indication when the routine is complete. In one
> > >> particular
> > >> > case, I am taking a 100 readings with NLPC set for 1000 so its a
> long
> > >> while
> > >> > before it's complete, but i have to guess when it's done.
> > >> >
> > >> > Thanks,
> > >> >
> > >> > Randy
> > >> >
> > >> >
> > >> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <
> randyevans2688@gmail.com>
> > >> > wrote:
> > >> >
> > >> > > Bill,
> > >> > >
> > >> > > I am trying to figure out the MATH function without much
> success. I
> > >> input
> > >> > > the sequence you said (I looked up the instructions to understand
> > >> what you
> > >> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
> > >> 4;TRIG; and
> > >> > > it shows up on the display when I input BLUE F1. I hit ENTER and
> it
> > >> takes
> > >> > > the 40 measurements and the MATH symbol shows on the display
> during
> > >> the
> > >> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2
> > >> and I
> > >> > > get a MATH ERR symbol on the display. I tried it a couple of
> times
> > >> and the
> > >> > > same result so I am doing something wrong. Is there a better
> source
> > >> for
> > >> > > explaining how to do front panel masurements than the User Guide,
> > >> which
> > >> > > seems oriented at programming automatic rather than manual
> > >> measurements.
> > >> > >
> > >> > > Randy
> > >> > >
> > >> > >
> > >> > >
> > >> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net>
> > >> wrote:
> > >> > >
> > >> > >> Randy:
> > >> > >>
> > >> > >> The MATH function is accessible from the keypad. I don't
> have
> > >> an IEEE
> > >> > >> interface right now that works. You can also program the numeric
> > >> keypad
> > >> > >> keys to have preprogrammed functions. DEFKEY
> > >> > >>
> > >> > >> I have made my own "low thermal" measurement leads from
> Pomona
> > >> #4892
> > >> > >> banana plugs and Belden #9272 wire. Why 9272, because it was
> handy
> > >> at the
> > >> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I
> have
> > >> plans
> > >> > >> to do custom cables with 16 ga. bare copper wire that I will
> twist
> > >> and
> > >> > >> then
> > >> > >> put a braided shield over it. I simply cannot find what I want
> so I
> > >> will
> > >> > >> build my own cable. I have done something like this before and
> it
> > >> worked
> > >> > >> fine. When I get a "round toit".
> > >> > >>
> > >> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that
> I
> > >> have
> > >> > >> used
> > >> > >> in the past to verify my homemade "low thermal" leads as
> described
> > >> above.
> > >> > >> Frankly I cannot see any difference between using the 1756 cables
> > >> and my
> > >> > >> homemade cables once I give them a few minutes for the thermals
> to go
> > >> > >> away.
> > >> > >> As far as I can tell and measure the differences, if any, are
> below
> > >> 0.1
> > >> > >> ppm
> > >> > >> at 10 volts.
> > >> > >>
> > >> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the
> 732A
> > >> are all
> > >> > >> adjustable you may be seeing a misadjusted 1 volt from the
> 732A. As
> > >> far
> > >> > >> as
> > >> > >> the instability of the readings it is hard to determine which is
> > >> causing
> > >> > >> the
> > >> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with
> the
> > >> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this
> > >> does is
> > >> > >> set
> > >> > >> the MATH to "Statistics" (store high reading/low reading/ and
> mean
> > >> of the
> > >> > >> readings) in the registers, the number of readings to "40", the
> > >> trigger to
> > >> > >> "hold" (which keeps the meter from triggering until I press
> "ENTER"
> > >> and
> > >> > >> then
> > >> > >> trigger the sequence of 40 readings when I push the "ENTER"
> button.
> > >> You
> > >> > >> can
> > >> > >> do all of this manually from the keypads but since I use this
> > >> sequence a
> > >> > >> lot
> > >> > >> I have preprogrammed it. This is after I set digits to "8" and
> PLC
> > >> to
> > >> > >> 100.
> > >> > >> Once those 40 readings are finished then you can access the
> various
> > >> MATH
> > >> > >> statistic registers, using the menu, by entering "MATH" and then
> a 2
> > >> for
> > >> > >> low, a 4 for mean, and 13 for high. Of course you could do all
> of
> > >> this
> > >> > >> through the IEEE also. The 3458A has a very rich set of
> measurement
> > >> > >> commands. I am still learning all of them. It depends upon
> what I
> > >> am
> > >> > >> trying to accomplish.
> > >> > >>
> > >> > >> Since the 1.018 and 1.0 volt outputs are passive and derived
> from
> > >> > >> resistive dividers from the 10 volt, I don't see how they could
> > >> contribute
> > >> > >> to the varying readings you are measuring. I think I would put a
> > >> short on
> > >> > >> the input of the 3458A and manually set the range to 1 volt and
> then
> > >> > >> observe
> > >> > >> the variations that way without the 732A involved. When I do
> this I
> > >> see a
> > >> > >> variation from low reading to high reading of 0.125 uVolts and
> then
> > >> > >> another
> > >> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to
> the
> > >> low
> > >> > >> side
> > >> > >> of the measurment terminals, GUARD connected doesn't seem to
> affect
> > >> the
> > >> > >> readings. So that is the base noise of the 3458A without the
> 732A,
> > >> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt
> > >> output I
> > >> > >> got
> > >> > >> a variation of 0.159 uVolts using the same 40 reading method
> above.
> > >> I
> > >> > >> would
> > >> > >> use this to determine where your problem might exist. Just
> having
> > >> the
> > >> > >> meter
> > >> > >> input shorted will point you in the right direction. Meter,
> cables
> > >> or
> > >> > >> 732A.
> > >> > >>
> > >> > >> Sorry for the long dissertation. Friends get mad at me for
> > >> being so
> > >> > >> detailed sometimes.
> > >> > >>
> > >> > >> Bill
> > >> > >>
> > >> > >> ----- Original Message -----
> > >> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > >> To: "Discussion of precise voltage measurement" <
> volt-nuts@febo.com>
> > >> > >> Sent: Sunday, August 24, 2014 7:22 AM
> > >> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >>
> > >> > >>
> > >> > >> > Bill,
> > >> > >> >
> > >> > >> > I have convinced myself that the problem I an seeing is due to
> > >> thermals.
> > >> > >> > If I move the cables (with gold-plated banana plugs) using a
> small
> > >> > >> towel
> > >> > >> > rather than letting my hand touch the plugs, it is much more
> > >> stable.
> > >> > >> If I
> > >> > >> > then hold the banana plug with my hand after the reading has
> > >> stabilized,
> > >> > >> > the reading drifts rapidly upward. I am trying to check the
> > >> stability
> > >> > >> of
> > >> > >> > the reading but I haven't figured out the MATH function yet. I
> > >> assume
> > >> > >> this
> > >> > >> > is a programmed function using GPIB only?
> > >> > >> >
> > >> > >> > The stability I am seeing by manually recording readings (using
> > >> NLPC of
> > >> > >> 100
> > >> > >> > and 1000) is much greater than what you are measuring on your
> > >> system.
> > >> > >> Not
> > >> > >> > sure how to ascertain if it's the 3458A or the 732. The value
> of
> > >> the
> > >> > >> > readings are very different between the two - the 3458 reads
> about
> > >> 50 uV
> > >> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> > >> Rather
> > >> > >> large
> > >> > >> > differences (this is after an ACAL). I need to find some
> better
> > >> cables
> > >> > >> to
> > >> > >> > make sure the errors are not due to thermals again.
> > >> > >> >
> > >> > >> > Randy
> > >> > >> >
> > >> > >> >
> > >> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <
> wpgold3637@att.net>
> > >> wrote:
> > >> > >> >
> > >> > >> > > Randy:
> > >> > >> > >
> > >> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they
> > >> will fit
> > >> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic
> > >> LC-R064R5C and
> > >> > >> > > others
> > >> > >> > > that are in this size and package. Order from one of the
> usual
> > >> > >> electronics
> > >> > >> > > distributors like Allied, Mouser, Digikey. This is a very
> common
> > >> > >> battery
> > >> > >> > > as
> > >> > >> > > it is used in a lot of "EXIT" signs so they are lighted when
> the
> > >> power
> > >> > >> goes
> > >> > >> > > out. I don't see how the 12V 7AH will fit as they are too
> > >> large. I
> > >> > >> guess
> > >> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2
> x 6
> > >> volt
> > >> > >> 4
> > >> > >> AH
> > >> > >> > > but the terminals are in the wrong place so you will have to
> > >> "nibble"
> > >> > >> out
> > >> > >> > > the aluminum plate that holds them in the 732A battery pack.
> > >> You have
> > >> > >> to
> > >> > >> > > be
> > >> > >> > > careful if you use the 12v 5AH as you will have 4 extra
> battery
> > >> > >> connection
> > >> > >> > > leads to deal with and connect correctly. I would stick with
> > >> the 6V
> > >> > >> 4AH.
> > >> > >> > > New batteries will last around 12 to 14 hours before the
> "CAL"
> > >> light
> > >> > >> goes
> > >> > >> > > out when AC power is not applied. So shipping to Cal Lab
> can be
> > >> a
> > >> > >> problem
> > >> > >> > > if it is a distance away, or you have to use a shipper like
> UPS
> > >> or
> > >> > >> FEDEX
> > >> > >> > > and
> > >> > >> > > you ship the night before and then use their "Morning
> delivery"
> > >> and
> > >> > >> the
> > >> > >> Cal
> > >> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> > >> course
> > >> > >> you
> > >> > >> > > could always strap another battery on the 732A and hook it
> up to
> > >> the
> > >> > >> "ext
> > >> > >> > > power" plug to last longer. I have seen it done. The issue
> is
> > >> to get
> > >> > >> the
> > >> > >> > > Cal Lab to charge the extra battery before they ship the 732A
> > >> back to
> > >> > >> you.
> > >> > >> > >
> > >> > >> > > When you remove and work on the battery pack always have
> the
> > >> AC
> > >> > >> power
> > >> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> > >> regulated
> > >> > >> supply
> > >> > >> > > is working.
> > >> > >> > >
> > >> > >> > > The "CAL" light is to indicate that power has not been
> lost
> > >> to the
> > >> > >> > > Reference Amp or other associated circuits. When the raw
> supply
> > >> > >> (battery)
> > >> > >> > > voltage drops below about 21 volts the "CAL" light will go
> out.
> > >> Below
> > >> > >> that
> > >> > >> > > voltage the heater circuits will not work correctly and the
> 18.6
> > >> volt
> > >> > >> > > regulated supply will not regulate. The requirement is that
> the
> > >> > >> Reference
> > >> > >> > > Amp be kept "alive" at all times to maintain the output
> voltage
> > >> that
> > >> > >> was
> > >> > >> > > measured at the time of the most recent Calibration or
> > >> Certification.
> > >> > >> When
> > >> > >> > > the semiconductor junctions are unbiased and cool off when
> power
> > >> is
> > >> > >> lost,
> > >> > >> > > and then power is restored the result will be a different 10
> > >> volts
> > >> > >> than
> > >> > >> > > before the power failure. My experience is that after all
> of the
> > >> > >> years
> > >> > >> > > that
> > >> > >> > > these units have been powered up, this won't happen and when
> > >> power is
> > >> > >> lost
> > >> > >> > > and then restored, even months later, the 732A will come
> back to
> > >> > >> almost
> > >> > >> > > exactly the same 10 volts as when they lost power, usually
> with
> > >> in 0.2
> > >> > >> PPM
> > >> > >> > > after 24 hours of "warm up".
> > >> > >> > >
> > >> > >> > > What type of hookup leads are you using when measuring
> the 1
> > >> volt
> > >> > >> > > output? What is the PLC set to? I always use 100 PLC to
> measure
> > >> > >> this.
> > >> > >> If
> > >> > >> > > you don't have "low thermal" connection leads you can
> experience
> > >> uV
> > >> > >> changes
> > >> > >> > > for a minute or more after plugging in the leads due to the
> > >> "thermals"
> > >> > >> > > generated because of the difference in temperature between
> the
> > >> banana
> > >> > >> jacks
> > >> > >> > > on the 732A and the banana plugs of the leads. I have found
> > >> that even
> > >> > >> just
> > >> > >> > > plugging in the lead will generate a thermal difference
> because
> > >> of
> > >> > >> > > difference of temps and some heating due to the physical act
> of
> > >> just
> > >> > >> > > inserting the plug because of friction between the jack and
> plug
> > >> (my
> > >> > >> theory
> > >> > >> > > at any rate). You have to allow at least a minute or more
> before
> > >> > >> being
> > >> > >> > > able
> > >> > >> > > to make a measurement after plugging in the leads. I just
> > >> measured
> > >> > >> the
> > >> > >> > > variation of the 1 volt output of my 732A and using my 3458A
> and
> > >> I
> > >> > >> got a
> > >> > >> > > total difference of 0.159 uV over 40 measurements using 100
> PLC
> > >> on
> > >> > >> the
> > >> > >> 1
> > >> > >> > > volt range of the 3458A. Using the MATH function and all of
> the
> > >> data
> > >> > >> you
> > >> > >> > > can collect. That was after waiting for several minutes
> after
> > >> > >> plugging
> > >> > >> in
> > >> > >> > > the leads.
> > >> > >> > >
> > >> > >> > > I hope all of this helps.
> > >> > >> > >
> > >> > >> > > Bill
> > >> > >> > >
> > >> > >> > >
> > >> > >> > >
> > >> > >> > > ----- Original Message -----
> > >> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > >> > > To: "Discussion of precise voltage measurement" <
> > >> volt-nuts@febo.com>
> > >> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> > >> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >> > >
> > >> > >> > >
> > >> > >> > > > Todd,
> > >> > >> > > >
> > >> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> > >> batteries
> > >> > >> that
> > >> > >> I
> > >> > >> > > > keep topped off and they have very low current draw (~2 to
> 3
> > >> mA at
> > >> > >> 13.5
> > >> > >> > > > VDC) when charged and at their float voltage, so I am
> pretty
> > >> sure
> > >> > >> they
> > >> > >> > > are
> > >> > >> > > > in good condition. I will look at getting those in the
> units
> > >> after
> > >> > >> I
> > >> > >> > > > ascertain the condition of the 732.
> > >> > >> > > >
> > >> > >> > > > So now I have a what appears to be a functioning 3458A and
> a
> > >> 732A
> > >> > >> but
> > >> > >> > > they
> > >> > >> > > > slightly disagree. I am like the man with two watches that
> > >> disagree
> > >> > >> on
> > >> > >> > > the
> > >> > >> > > > time - which is correct? For the moment, i am only
> concerned
> > >> with
> > >> > >> > > > stability. The need for absolute accuracy will come later.
> > >> > >> > > >
> > >> > >> > > > Randy
> > >> > >> > > >
> > >> > >> > > >
> > >> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <
> > >> tmicallef@gmail.com
> > >> > >> >
> > >> > >> > > wrote:
> > >> > >> > > >
> > >> > >> > > > > Randy,
> > >> > >> > > > >
> > >> > >> > > > > You have two possible choices. It can be configured with
> 4 x
> > >> 6v
> > >> > >> 4Ah
> > >> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the
> previous
> > >> owner
> > >> > >> has
> > >> > >> > > > > modified the battery pack already. A couple of mine
> needed a
> > >> > >> nibbler
> > >> > >> > > tool
> > >> > >> > > > > to remove enough of the aluminum cover that fits over the
> > >> tops of
> > >> > >> the
> > >> > >> > > > > batteries. The original cover will short out to the
> battery
> > >> tabs
> > >> > >> > > regardless
> > >> > >> > > > > of the battery configuration if this is not done.
> > >> > >> > > > >
> > >> > >> > > > > You can find larger capacity batteries that will give you
> > >> slightly
> > >> > >> more
> > >> > >> > > > > battery life. I lost a couple sets of mail-order
> batteries
> > >> after a
> > >> > >> few
> > >> > >> > > > > extended outages. I would recommend going with locally
> bought
> > >> > >> batteries
> > >> > >> > > > > instead of the cheaper mail order. My local Batteries
> Plus
> > >> will
> > >> > >> > > typically
> > >> > >> > > > > have some warranty if I remember correctly. Moving
> forward I
> > >> will
> > >> > >> only
> > >> > >> > > use
> > >> > >> > > > > 2 12v batteries and pre-charge them on a battery charger
> to
> > >> > >> equalize
> > >> > >> > > them
> > >> > >> > > > > before putting them in the 732A. I think the cheap
> batteries
> > >> did
> > >> > >> not
> > >> > >> > > > > discharge equally, and would not recover when power was
> > >> applied.
> > >> > >> > > > >
> > >> > >> > > > > Inspect the back plane for damaged traces and look at the
> > >> > >> capacitors. I
> > >> > >> > > had
> > >> > >> > > > > a few that looked questionable. So far, I have replaced
> all
> > >> the
> > >> > >> big
> > >> > >> > > caps
> > >> > >> > > on
> > >> > >> > > > > the pre-regulator and regulator boards. My feeling is
> that
> > >> once
> > >> > >> these
> > >> > >> > > go
> > >> > >> > > > > online, they should run as long as possible between
> repairs.
> > >> > >> > > > >
> > >> > >> > > > > The battery charger circuit may need adjusting. I tweaked
> > >> mine and
> > >> > >> it
> > >> > >> > > > > seemed to work fine.
> > >> > >> > > > >
> > >> > >> > > > > Todd
> > >> > >> > > > >
> > >> > >> > > > >
> > >> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > >> > >> > > randyevans2688@gmail.com>
> > >> > >> > > > > wrote:
> > >> > >> > > > >
> > >> > >> > > > > > I received my Fluke 732A today. Just powered it up
> but it
> > >> needs
> > >> > >> new
> > >> > >> > > > > > batteries. Any suggestions for sources (I haven't
> opened
> > >> up the
> > >> > >> unit
> > >> > >> > > > > yet -
> > >> > >> > > > > > I want to make sure it works before doing that). Also
> > >> received
> > >> > >> the
> > >> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims'
> CAL
> > >> ran
> > >> > >> data
> > >> > >> > > > > dumper
> > >> > >> > > > > > program to get the CAL data from my 3458A. Should be a
> > >> busy
> > >> > >> weekend.
> > >> > >> > > > > >
> > >> > >> > > > > > Randy
> > >> > >> > > > > > _______________________________________________
> > >> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > > > > To unsubscribe, go to
> > >> > >> > > > > >
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > > > > and follow the instructions there.
> > >> > >> > > > > >
> > >> > >> > > > > _______________________________________________
> > >> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > > > To unsubscribe, go to
> > >> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > > > and follow the instructions there.
> > >> > >> > > > >
> > >> > >> > > > _______________________________________________
> > >> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > > To unsubscribe, go to
> > >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > > and follow the instructions there.
> > >> > >> > >
> > >> > >> > > _______________________________________________
> > >> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > > To unsubscribe, go to
> > >> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > > and follow the instructions there.
> > >> > >> > >
> > >> > >> > _______________________________________________
> > >> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> > To unsubscribe, go to
> > >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> > and follow the instructions there.
> > >> > >>
> > >> > >> _______________________________________________
> > >> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> > >> To unsubscribe, go to
> > >> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > >> and follow the instructions there.
> > >> > >>
> > >> > >
> > >> > >
> > >> > _______________________________________________
> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > and follow the instructions there.
> > >> _______________________________________________
> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> and follow the instructions there.
> > >>
> > >
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>