D
Don@True-Cal
Tue, Aug 26, 2014 1:54 AM
Why?
Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
Silver or Gold plating on the terminal or wire will introduce the
undesirable dissimilar metal properties, both at the plating junction
and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
--
Mike
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Why?
Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
> Silver or Gold plating on the terminal or wire will introduce the
> undesirable dissimilar metal properties, both at the plating junction
> and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
--
Mike
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
MS
Mike S
Tue, Aug 26, 2014 4:07 AM
On 8/25/2014 9:54 PM, Don@True-Cal wrote:
Why?
Let me count the ways.You can never count on any Seebeck voltage to
be immediately offset, there are far too many variables.
A set of
5440A-7002 (banana plug) cables comes with this calibrator
(5440A-7003 spade lugs for 5720A)
For those leads, Fluke says they have the advantage of "minimizing the
emf caused by temperature difference between the ends of the leads." I
don't have much argument with that - if there's a temperature difference
between the ends, you won't have an equal thermocouple offset, so
minimizing such effects makes sense.
But, since those leads use pure copper wiring ("Belden 8262 or 8719 or
equivalent"), how does having only the contact itself made of tellurium
copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas
tight crimped connections? Does the tinning on that copper cable have an
effect? What's the Seebeck coefficient for a copper to tellurium copper
connection?
You said that "Tellurium Copper is usually not used for a device's
terminal posts but used as the lead wire..." But, Fluke does exactly the
opposite - those Fluke leads use TC connectors and are characterized
"while engaged in a five-way binding post of Tellurium Copper Alloy 145,
half hard." Since they use pure copper cabling, I'd think they'd work
even better with a pure copper binding post, since the offsetting
thermocouples would be in close proximity, and not at different ends of
the cable.
You earlier said "Any type of Silver or Gold plating on the terminal or
wire will introduce the undesirable dissimilar metal properties, both
at the plating junction and at the plating metal to DUT terminal." Now
you emphasize that by saying "If someone used a set of gold-plated
interconnects for this procedure, they would be laughed out of the lab
and the calibrator would be useless until recalibrated properly..."
But, the specs for the Fluke 5440A-7003 which you say are appropriate
for the 5720A state: "Connector materials: Copper space lug with gold
plating over silver plating." (Both platings!!) Is someone to be laughed
out of the lab for using the cables they're supposed to use? The
5440A-7002 specs aren't clear, but if they're not plated, are you
required to clean them of copper oxide for every use? It seems that
would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while
a copper-copper oxide one is ~1000 uV/C.
You seem to be stating one thing, but then giving evidence which
contradicts your claim. It seems that Fluke uses TC where there's
thermal mass, and pure copper where there's little, and uses plating in
addition.
Please clarify, this isn't making sense to me.
On 8/25/2014 9:54 PM, Don@True-Cal wrote:
> Why?
>
> Let me count the ways.You can never count on any Seebeck voltage to
> be immediately offset, there are far too many variables.
...
> A set of
> 5440A-7002 (banana plug) cables comes with this calibrator
> (5440A-7003 spade lugs for 5720A)
For those leads, Fluke says they have the advantage of "minimizing the
emf caused by temperature difference between the ends of the leads." I
don't have much argument with that - if there's a temperature difference
between the ends, you won't have an equal thermocouple offset, so
minimizing such effects makes sense.
But, since those leads use pure copper wiring ("Belden 8262 or 8719 or
equivalent"), how does having only the contact itself made of tellurium
copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas
tight crimped connections? Does the tinning on that copper cable have an
effect? What's the Seebeck coefficient for a copper to tellurium copper
connection?
You said that "Tellurium Copper is usually not used for a device's
terminal posts but used as the lead wire..." But, Fluke does exactly the
opposite - those Fluke leads use TC connectors and are characterized
"while engaged in a five-way binding post of Tellurium Copper Alloy 145,
half hard." Since they use pure copper cabling, I'd think they'd work
even better with a pure copper binding post, since the offsetting
thermocouples would be in close proximity, and not at different ends of
the cable.
You earlier said "Any type of Silver or Gold plating on the terminal or
wire will introduce the undesirable dissimilar metal properties, both
at the plating junction and at the plating metal to DUT terminal." Now
you emphasize that by saying "If someone used a set of gold-plated
interconnects for this procedure, they would be laughed out of the lab
and the calibrator would be useless until recalibrated properly..."
But, the specs for the Fluke 5440A-7003 which you say are appropriate
for the 5720A state: "Connector materials: Copper space lug with gold
plating over silver plating." (Both platings!!) Is someone to be laughed
out of the lab for using the cables they're supposed to use? The
5440A-7002 specs aren't clear, but if they're not plated, are you
required to clean them of copper oxide for every use? It seems that
would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while
a copper-copper oxide one is ~1000 uV/C.
You seem to be stating one thing, but then giving evidence which
contradicts your claim. It seems that Fluke uses TC where there's
thermal mass, and pure copper where there's little, and uses plating in
addition.
Please clarify, this isn't making sense to me.
RE
Randy Evans
Tue, Aug 26, 2014 4:23 AM
Is there any way to tell when the function key routine is complete? In the
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long while
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1. I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and the
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an IEEE
interface right now that works. You can also program the numeric keypad
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona #4892
banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I will
build my own cable. I have done something like this before and it worked
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
as
the instability of the readings it is hard to determine which is causing
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button. You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring. I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being so
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward. I am trying to check the stability
the reading but I haven't figured out the MATH function yet. I assume
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC of
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better cables
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the power
out. I don't see how the 12V 7AH will fit as they are too large. I
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt
but the terminals are in the wrong place so you will have to "nibble"
the aluminum plate that holds them in the 732A battery pack. You have
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the 6V
New batteries will last around 12 to 14 hours before the "CAL" light
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of course
could always strap another battery on the 732A and hook it up to the
power" plug to last longer. I have seen it done. The issue is to get
Cal Lab to charge the extra battery before they ship the 732A back to
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v regulated
is working.
The "CAL" light is to indicate that power has not been lost to the
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out. Below
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage that
measured at the time of the most recent Calibration or Certification.
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power is
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in 0.2
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1 volt
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana
on the 732A and the banana plugs of the leads. I have found that even
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the data
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA at
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units after
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that disagree
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops of
batteries. The original cover will short out to the battery tabs
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you slightly
battery life. I lost a couple sets of mail-order batteries after a
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I will
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries did
discharge equally, and would not recover when power was applied.
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine and
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it needs
batteries. Any suggestions for sources (I haven't opened up the
I want to make sure it works before doing that). Also received
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
Is there any way to tell when the function key routine is complete? In the
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long while
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
wrote:
> Bill,
>
> I am trying to figure out the MATH function without much success. I input
> the sequence you said (I looked up the instructions to understand what you
> did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
> it shows up on the display when I input BLUE F1. I hit ENTER and it takes
> the 40 measurements and the MATH symbol shows on the display during the
> measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> get a MATH ERR symbol on the display. I tried it a couple of times and the
> same result so I am doing something wrong. Is there a better source for
> explaining how to do front panel masurements than the User Guide, which
> seems oriented at programming automatic rather than manual measurements.
>
> Randy
>
>
>
> On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
>
>> Randy:
>>
>> The MATH function is accessible from the keypad. I don't have an IEEE
>> interface right now that works. You can also program the numeric keypad
>> keys to have preprogrammed functions. DEFKEY
>>
>> I have made my own "low thermal" measurement leads from Pomona #4892
>> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
>> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
>> to do custom cables with 16 ga. bare copper wire that I will twist and
>> then
>> put a braided shield over it. I simply cannot find what I want so I will
>> build my own cable. I have done something like this before and it worked
>> fine. When I get a "round toit".
>>
>> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
>> used
>> in the past to verify my homemade "low thermal" leads as described above.
>> Frankly I cannot see any difference between using the 1756 cables and my
>> homemade cables once I give them a few minutes for the thermals to go
>> away.
>> As far as I can tell and measure the differences, if any, are below 0.1
>> ppm
>> at 10 volts.
>>
>> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
>> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far
>> as
>> the instability of the readings it is hard to determine which is causing
>> the
>> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
>> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
>> set
>> the MATH to "Statistics" (store high reading/low reading/ and mean of the
>> readings) in the registers, the number of readings to "40", the trigger to
>> "hold" (which keeps the meter from triggering until I press "ENTER" and
>> then
>> trigger the sequence of 40 readings when I push the "ENTER" button. You
>> can
>> do all of this manually from the keypads but since I use this sequence a
>> lot
>> I have preprogrammed it. This is after I set digits to "8" and PLC to
>> 100.
>> Once those 40 readings are finished then you can access the various MATH
>> statistic registers, using the menu, by entering "MATH" and then a 2 for
>> low, a 4 for mean, and 13 for high. Of course you could do all of this
>> through the IEEE also. The 3458A has a very rich set of measurement
>> commands. I am still learning all of them. It depends upon what I am
>> trying to accomplish.
>>
>> Since the 1.018 and 1.0 volt outputs are passive and derived from
>> resistive dividers from the 10 volt, I don't see how they could contribute
>> to the varying readings you are measuring. I think I would put a short on
>> the input of the 3458A and manually set the range to 1 volt and then
>> observe
>> the variations that way without the 732A involved. When I do this I see a
>> variation from low reading to high reading of 0.125 uVolts and then
>> another
>> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
>> side
>> of the measurment terminals, GUARD connected doesn't seem to affect the
>> readings. So that is the base noise of the 3458A without the 732A,
>> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
>> got
>> a variation of 0.159 uVolts using the same 40 reading method above. I
>> would
>> use this to determine where your problem might exist. Just having the
>> meter
>> input shorted will point you in the right direction. Meter, cables or
>> 732A.
>>
>> Sorry for the long dissertation. Friends get mad at me for being so
>> detailed sometimes.
>>
>> Bill
>>
>> ----- Original Message -----
>> From: "Randy Evans" <randyevans2688@gmail.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Sent: Sunday, August 24, 2014 7:22 AM
>> Subject: Re: [volt-nuts] 732A and Prologix received
>>
>>
>> > Bill,
>> >
>> > I have convinced myself that the problem I an seeing is due to thermals.
>> > If I move the cables (with gold-plated banana plugs) using a small
>> towel
>> > rather than letting my hand touch the plugs, it is much more stable.
>> If I
>> > then hold the banana plug with my hand after the reading has stabilized,
>> > the reading drifts rapidly upward. I am trying to check the stability
>> of
>> > the reading but I haven't figured out the MATH function yet. I assume
>> this
>> > is a programmed function using GPIB only?
>> >
>> > The stability I am seeing by manually recording readings (using NLPC of
>> 100
>> > and 1000) is much greater than what you are measuring on your system.
>> Not
>> > sure how to ascertain if it's the 3458A or the 732. The value of the
>> > readings are very different between the two - the 3458 reads about 50 uV
>> > high on the 10 V output and about 12 uV low on the 1V output. Rather
>> large
>> > differences (this is after an ACAL). I need to find some better cables
>> to
>> > make sure the errors are not due to thermals again.
>> >
>> > Randy
>> >
>> >
>> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote:
>> >
>> > > Randy:
>> > >
>> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
>> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
>> > > others
>> > > that are in this size and package. Order from one of the usual
>> electronics
>> > > distributors like Allied, Mouser, Digikey. This is a very common
>> battery
>> > > as
>> > > it is used in a lot of "EXIT" signs so they are lighted when the power
>> goes
>> > > out. I don't see how the 12V 7AH will fit as they are too large. I
>> guess
>> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt
>> 4
>> AH
>> > > but the terminals are in the wrong place so you will have to "nibble"
>> out
>> > > the aluminum plate that holds them in the 732A battery pack. You have
>> to
>> > > be
>> > > careful if you use the 12v 5AH as you will have 4 extra battery
>> connection
>> > > leads to deal with and connect correctly. I would stick with the 6V
>> 4AH.
>> > > New batteries will last around 12 to 14 hours before the "CAL" light
>> goes
>> > > out when AC power is not applied. So shipping to Cal Lab can be a
>> problem
>> > > if it is a distance away, or you have to use a shipper like UPS or
>> FEDEX
>> > > and
>> > > you ship the night before and then use their "Morning delivery" and
>> the
>> Cal
>> > > Lab is expecting your 732A. Same on the way back to you. Of course
>> you
>> > > could always strap another battery on the 732A and hook it up to the
>> "ext
>> > > power" plug to last longer. I have seen it done. The issue is to get
>> the
>> > > Cal Lab to charge the extra battery before they ship the 732A back to
>> you.
>> > >
>> > > When you remove and work on the battery pack always have the AC
>> power
>> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
>> supply
>> > > is working.
>> > >
>> > > The "CAL" light is to indicate that power has not been lost to the
>> > > Reference Amp or other associated circuits. When the raw supply
>> (battery)
>> > > voltage drops below about 21 volts the "CAL" light will go out. Below
>> that
>> > > voltage the heater circuits will not work correctly and the 18.6 volt
>> > > regulated supply will not regulate. The requirement is that the
>> Reference
>> > > Amp be kept "alive" at all times to maintain the output voltage that
>> was
>> > > measured at the time of the most recent Calibration or Certification.
>> When
>> > > the semiconductor junctions are unbiased and cool off when power is
>> lost,
>> > > and then power is restored the result will be a different 10 volts
>> than
>> > > before the power failure. My experience is that after all of the
>> years
>> > > that
>> > > these units have been powered up, this won't happen and when power is
>> lost
>> > > and then restored, even months later, the 732A will come back to
>> almost
>> > > exactly the same 10 volts as when they lost power, usually with in 0.2
>> PPM
>> > > after 24 hours of "warm up".
>> > >
>> > > What type of hookup leads are you using when measuring the 1 volt
>> > > output? What is the PLC set to? I always use 100 PLC to measure
>> this.
>> If
>> > > you don't have "low thermal" connection leads you can experience uV
>> changes
>> > > for a minute or more after plugging in the leads due to the "thermals"
>> > > generated because of the difference in temperature between the banana
>> jacks
>> > > on the 732A and the banana plugs of the leads. I have found that even
>> just
>> > > plugging in the lead will generate a thermal difference because of
>> > > difference of temps and some heating due to the physical act of just
>> > > inserting the plug because of friction between the jack and plug (my
>> theory
>> > > at any rate). You have to allow at least a minute or more before
>> being
>> > > able
>> > > to make a measurement after plugging in the leads. I just measured
>> the
>> > > variation of the 1 volt output of my 732A and using my 3458A and I
>> got a
>> > > total difference of 0.159 uV over 40 measurements using 100 PLC on
>> the
>> 1
>> > > volt range of the 3458A. Using the MATH function and all of the data
>> you
>> > > can collect. That was after waiting for several minutes after
>> plugging
>> in
>> > > the leads.
>> > >
>> > > I hope all of this helps.
>> > >
>> > > Bill
>> > >
>> > >
>> > >
>> > > ----- Original Message -----
>> > > From: "Randy Evans" <randyevans2688@gmail.com>
>> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> > > Sent: Friday, August 22, 2014 9:03 PM
>> > > Subject: Re: [volt-nuts] 732A and Prologix received
>> > >
>> > >
>> > > > Todd,
>> > > >
>> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
>> that
>> I
>> > > > keep topped off and they have very low current draw (~2 to 3 mA at
>> 13.5
>> > > > VDC) when charged and at their float voltage, so I am pretty sure
>> they
>> > > are
>> > > > in good condition. I will look at getting those in the units after
>> I
>> > > > ascertain the condition of the 732.
>> > > >
>> > > > So now I have a what appears to be a functioning 3458A and a 732A
>> but
>> > > they
>> > > > slightly disagree. I am like the man with two watches that disagree
>> on
>> > > the
>> > > > time - which is correct? For the moment, i am only concerned with
>> > > > stability. The need for absolute accuracy will come later.
>> > > >
>> > > > Randy
>> > > >
>> > > >
>> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com
>> >
>> > > wrote:
>> > > >
>> > > > > Randy,
>> > > > >
>> > > > > You have two possible choices. It can be configured with 4 x 6v
>> 4Ah
>> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
>> has
>> > > > > modified the battery pack already. A couple of mine needed a
>> nibbler
>> > > tool
>> > > > > to remove enough of the aluminum cover that fits over the tops of
>> the
>> > > > > batteries. The original cover will short out to the battery tabs
>> > > regardless
>> > > > > of the battery configuration if this is not done.
>> > > > >
>> > > > > You can find larger capacity batteries that will give you slightly
>> more
>> > > > > battery life. I lost a couple sets of mail-order batteries after a
>> few
>> > > > > extended outages. I would recommend going with locally bought
>> batteries
>> > > > > instead of the cheaper mail order. My local Batteries Plus will
>> > > typically
>> > > > > have some warranty if I remember correctly. Moving forward I will
>> only
>> > > use
>> > > > > 2 12v batteries and pre-charge them on a battery charger to
>> equalize
>> > > them
>> > > > > before putting them in the 732A. I think the cheap batteries did
>> not
>> > > > > discharge equally, and would not recover when power was applied.
>> > > > >
>> > > > > Inspect the back plane for damaged traces and look at the
>> capacitors. I
>> > > had
>> > > > > a few that looked questionable. So far, I have replaced all the
>> big
>> > > caps
>> > > on
>> > > > > the pre-regulator and regulator boards. My feeling is that once
>> these
>> > > go
>> > > > > online, they should run as long as possible between repairs.
>> > > > >
>> > > > > The battery charger circuit may need adjusting. I tweaked mine and
>> it
>> > > > > seemed to work fine.
>> > > > >
>> > > > > Todd
>> > > > >
>> > > > >
>> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
>> > > randyevans2688@gmail.com>
>> > > > > wrote:
>> > > > >
>> > > > > > I received my Fluke 732A today. Just powered it up but it needs
>> new
>> > > > > > batteries. Any suggestions for sources (I haven't opened up the
>> unit
>> > > > > yet -
>> > > > > > I want to make sure it works before doing that). Also received
>> the
>> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
>> data
>> > > > > dumper
>> > > > > > program to get the CAL data from my 3458A. Should be a busy
>> weekend.
>> > > > > >
>> > > > > > Randy
>> > > > > > _______________________________________________
>> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > > > To unsubscribe, go to
>> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > > > and follow the instructions there.
>> > > > > >
>> > > > > _______________________________________________
>> > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > > To unsubscribe, go to
>> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > > and follow the instructions there.
>> > > > >
>> > > > _______________________________________________
>> > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > > To unsubscribe, go to
>> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > > and follow the instructions there.
>> > >
>> > > _______________________________________________
>> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > To unsubscribe, go to
>> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > > and follow the instructions there.
>> > >
>> > _______________________________________________
>> > volt-nuts mailing list -- volt-nuts@febo.com
>> > To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> > and follow the instructions there.
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
>
CB
Charles Black
Tue, Aug 26, 2014 4:28 AM
My post on this thread earlier today seems to be lost (maybe the system
change) so I am re-posting some of it (corrected) since it's apropos to
the discussion.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire. This is very convenient since it is so
easy to duplicate in the field and makes simple inexpensive test leads
best for high precision measurements. In order for the 3458A to make
full accuracy measurements (8 digit) NPLC must be set to 1000 (according
to the User's Guide). If you use a lower NPLC value there is a table in
the User's Guide that can be used to determine how accurate your
measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv ± 10nV Equilibration time 5
minutes. 14 gauge per Calibration Manual.
"U" heavy wire -0.00021mv ± 10nV Equil. time 5 minutes.
Used the through holes in the Input banana posts only.
Copper wire -0.00019mv ± 10nV NAPA PVC covered
automobile wire at same contact points as CAL shunt
Standard Ground Plate -0.00040mv ± 10nV Equil. time 5 minutes. Gold
plated ground plate from my Datron 4910
Copper wire -0.00019mv ± 10nv Equil. time 2
seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv ± 10nV Equil. time 2 seconds.
M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv ± 30nV Equil. time 20
minutes. My best "Perfect" gold plated plugs with copper wire.
Note that all my test shorts equilibrate to virtualy zero volts. There
is almost no Seebeck voltage for any of my test shorts as long as the
system is allowed to equilibrate. Equilibation times varied between less
than 2 seconds up 20 minutes with longer times due to excessive thermal
mass. Excessive thermal mass also caused some voltage instability.
Charlie
On 8/25/2014 6:54 PM, Don@True-Cal wrote:
Why?
Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
Don Johnson
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received
On 8/25/2014 11:02 AM, Don@True-Cal wrote:
Silver or Gold plating on the terminal or wire will introduce the
undesirable dissimilar metal properties, both at the plating junction
and at the plating metal to DUT terminal.
Why?
Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.
Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
--
Mike
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
My post on this thread earlier today seems to be lost (maybe the system
change) so I am re-posting some of it (corrected) since it's apropos to
the discussion.
Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire. This is very convenient since it is so
easy to duplicate in the field and makes simple inexpensive test leads
best for high precision measurements. In order for the 3458A to make
full accuracy measurements (8 digit) NPLC must be set to 1000 (according
to the User's Guide). If you use a lower NPLC value there is a table in
the User's Guide that can be used to determine how accurate your
measurements are going to be.
Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.
Calibration shunt -0.00021mv ± 10nV Equilibration time 5
minutes. 14 gauge per Calibration Manual.
"U" heavy wire -0.00021mv ± 10nV Equil. time 5 minutes.
Used the through holes in the Input banana posts only.
Copper wire -0.00019mv ± 10nV NAPA PVC covered
automobile wire at same contact points as CAL shunt
Standard Ground Plate -0.00040mv ± 10nV Equil. time 5 minutes. Gold
plated ground plate from my Datron 4910
Copper wire -0.00019mv ± 10nv Equil. time 2
seconds. Used the banana through holes.
STP 2 meter test lead -0.00021mv ± 10nV Equil. time 2 seconds.
M27500 24 gauge STP Tefzel insulation.
Banana plugs -0.00021mv ± 30nV Equil. time 20
minutes. My best "Perfect" gold plated plugs with copper wire.
Note that all my test shorts equilibrate to virtualy zero volts. There
is almost no Seebeck voltage for any of my test shorts as long as the
system is allowed to equilibrate. Equilibation times varied between less
than 2 seconds up 20 minutes with longer times due to excessive thermal
mass. Excessive thermal mass also caused some voltage instability.
Charlie
On 8/25/2014 6:54 PM, Don@True-Cal wrote:
> Why?
>
> Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.
>
> BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.
>
> This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.
>
> Don Johnson
>
> -----Original Message-----
> From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
> Sent: Monday, August 25, 2014 11:03 AM
> To: volt-nuts@febo.com
> Subject: Re: [volt-nuts] 732A and Prologix received
>
> On 8/25/2014 11:02 AM, Don@True-Cal wrote:
>> Silver or Gold plating on the terminal or wire will introduce the
>> undesirable dissimilar metal properties, both at the plating junction
>> and at the plating metal to DUT terminal.
> Why?
>
> Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.
>
> A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
> shortly after plugging in a banana plug warmed by body heat), they simply cancel.
>
> Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.
>
> --
> Mike
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
>
BG
Bill Gold
Tue, Aug 26, 2014 5:01 AM
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front panel
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand what
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display. I tried it a couple of times and
same result so I am doing something wrong. Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.
Randy
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy at
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the trigger
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this sequence
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of this
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a short
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC
and 1000) is much greater than what you are measuring on your system.
sure how to ascertain if it's the 3458A or the 732. The value of the
readings are very different between the two - the 3458 reads about 50
high on the 10 V output and about 12 uV low on the 1V output. Rather
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too large.
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery" and
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is to
Cal Lab to charge the extra battery before they ship the 732A back
When you remove and work on the battery pack always have the AC
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost to
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power is
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience uV
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just measured
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC on
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH batteries
keep topped off and they have very low current draw (~2 to 3 mA
VDC) when charged and at their float voltage, so I am pretty sure
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a 732A
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4 x 6v
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus will
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all the
the pre-regulator and regulator boards. My feeling is that once
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened up
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front panel
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" <randyevans2688@gmail.com>
To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
> Is there any way to tell when the function key routine is complete? In
the
> case of taking multiple readings using the DEFKEY and MATH function, I
> don't see any indication when the routine is complete. In one particular
> case, I am taking a 100 readings with NLPC set for 1000 so its a long
while
> before it's complete, but i have to guess when it's done.
>
> Thanks,
>
> Randy
>
>
> On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > Bill,
> >
> > I am trying to figure out the MATH function without much success. I
input
> > the sequence you said (I looked up the instructions to understand what
you
> > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
and
> > it shows up on the display when I input BLUE F1. I hit ENTER and it
takes
> > the 40 measurements and the MATH symbol shows on the display during the
> > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I
> > get a MATH ERR symbol on the display. I tried it a couple of times and
the
> > same result so I am doing something wrong. Is there a better source for
> > explaining how to do front panel masurements than the User Guide, which
> > seems oriented at programming automatic rather than manual measurements.
> >
> > Randy
> >
> >
> >
> > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> >
> >> Randy:
> >>
> >> The MATH function is accessible from the keypad. I don't have an
IEEE
> >> interface right now that works. You can also program the numeric
keypad
> >> keys to have preprogrammed functions. DEFKEY
> >>
> >> I have made my own "low thermal" measurement leads from Pomona
#4892
> >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at
the
> >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
plans
> >> to do custom cables with 16 ga. bare copper wire that I will twist and
> >> then
> >> put a braided shield over it. I simply cannot find what I want so I
will
> >> build my own cable. I have done something like this before and it
worked
> >> fine. When I get a "round toit".
> >>
> >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> >> used
> >> in the past to verify my homemade "low thermal" leads as described
above.
> >> Frankly I cannot see any difference between using the 1756 cables and
my
> >> homemade cables once I give them a few minutes for the thermals to go
> >> away.
> >> As far as I can tell and measure the differences, if any, are below 0.1
> >> ppm
> >> at 10 volts.
> >>
> >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
all
> >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
far
> >> as
> >> the instability of the readings it is hard to determine which is
causing
> >> the
> >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
is
> >> set
> >> the MATH to "Statistics" (store high reading/low reading/ and mean of
the
> >> readings) in the registers, the number of readings to "40", the trigger
to
> >> "hold" (which keeps the meter from triggering until I press "ENTER" and
> >> then
> >> trigger the sequence of 40 readings when I push the "ENTER" button.
You
> >> can
> >> do all of this manually from the keypads but since I use this sequence
a
> >> lot
> >> I have preprogrammed it. This is after I set digits to "8" and PLC to
> >> 100.
> >> Once those 40 readings are finished then you can access the various
MATH
> >> statistic registers, using the menu, by entering "MATH" and then a 2
for
> >> low, a 4 for mean, and 13 for high. Of course you could do all of this
> >> through the IEEE also. The 3458A has a very rich set of measurement
> >> commands. I am still learning all of them. It depends upon what I am
> >> trying to accomplish.
> >>
> >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> >> resistive dividers from the 10 volt, I don't see how they could
contribute
> >> to the varying readings you are measuring. I think I would put a short
on
> >> the input of the 3458A and manually set the range to 1 volt and then
> >> observe
> >> the variations that way without the 732A involved. When I do this I
see a
> >> variation from low reading to high reading of 0.125 uVolts and then
> >> another
> >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
> >> side
> >> of the measurment terminals, GUARD connected doesn't seem to affect the
> >> readings. So that is the base noise of the 3458A without the 732A,
> >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I
> >> got
> >> a variation of 0.159 uVolts using the same 40 reading method above. I
> >> would
> >> use this to determine where your problem might exist. Just having the
> >> meter
> >> input shorted will point you in the right direction. Meter, cables or
> >> 732A.
> >>
> >> Sorry for the long dissertation. Friends get mad at me for being
so
> >> detailed sometimes.
> >>
> >> Bill
> >>
> >> ----- Original Message -----
> >> From: "Randy Evans" <randyevans2688@gmail.com>
> >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> Sent: Sunday, August 24, 2014 7:22 AM
> >> Subject: Re: [volt-nuts] 732A and Prologix received
> >>
> >>
> >> > Bill,
> >> >
> >> > I have convinced myself that the problem I an seeing is due to
thermals.
> >> > If I move the cables (with gold-plated banana plugs) using a small
> >> towel
> >> > rather than letting my hand touch the plugs, it is much more stable.
> >> If I
> >> > then hold the banana plug with my hand after the reading has
stabilized,
> >> > the reading drifts rapidly upward. I am trying to check the
stability
> >> of
> >> > the reading but I haven't figured out the MATH function yet. I
assume
> >> this
> >> > is a programmed function using GPIB only?
> >> >
> >> > The stability I am seeing by manually recording readings (using NLPC
of
> >> 100
> >> > and 1000) is much greater than what you are measuring on your system.
> >> Not
> >> > sure how to ascertain if it's the 3458A or the 732. The value of the
> >> > readings are very different between the two - the 3458 reads about 50
uV
> >> > high on the 10 V output and about 12 uV low on the 1V output. Rather
> >> large
> >> > differences (this is after an ACAL). I need to find some better
cables
> >> to
> >> > make sure the errors are not due to thermals again.
> >> >
> >> > Randy
> >> >
> >> >
> >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
wrote:
> >> >
> >> > > Randy:
> >> > >
> >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
fit
> >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
and
> >> > > others
> >> > > that are in this size and package. Order from one of the usual
> >> electronics
> >> > > distributors like Allied, Mouser, Digikey. This is a very common
> >> battery
> >> > > as
> >> > > it is used in a lot of "EXIT" signs so they are lighted when the
power
> >> goes
> >> > > out. I don't see how the 12V 7AH will fit as they are too large.
I
> >> guess
> >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
volt
> >> 4
> >> AH
> >> > > but the terminals are in the wrong place so you will have to
"nibble"
> >> out
> >> > > the aluminum plate that holds them in the 732A battery pack. You
have
> >> to
> >> > > be
> >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> >> connection
> >> > > leads to deal with and connect correctly. I would stick with the
6V
> >> 4AH.
> >> > > New batteries will last around 12 to 14 hours before the "CAL"
light
> >> goes
> >> > > out when AC power is not applied. So shipping to Cal Lab can be a
> >> problem
> >> > > if it is a distance away, or you have to use a shipper like UPS or
> >> FEDEX
> >> > > and
> >> > > you ship the night before and then use their "Morning delivery" and
> >> the
> >> Cal
> >> > > Lab is expecting your 732A. Same on the way back to you. Of
course
> >> you
> >> > > could always strap another battery on the 732A and hook it up to
the
> >> "ext
> >> > > power" plug to last longer. I have seen it done. The issue is to
get
> >> the
> >> > > Cal Lab to charge the extra battery before they ship the 732A back
to
> >> you.
> >> > >
> >> > > When you remove and work on the battery pack always have the AC
> >> power
> >> > > plugged in. The "CAL" led will stay on because the 18.6 v
regulated
> >> supply
> >> > > is working.
> >> > >
> >> > > The "CAL" light is to indicate that power has not been lost to
the
> >> > > Reference Amp or other associated circuits. When the raw supply
> >> (battery)
> >> > > voltage drops below about 21 volts the "CAL" light will go out.
Below
> >> that
> >> > > voltage the heater circuits will not work correctly and the 18.6
volt
> >> > > regulated supply will not regulate. The requirement is that the
> >> Reference
> >> > > Amp be kept "alive" at all times to maintain the output voltage
that
> >> was
> >> > > measured at the time of the most recent Calibration or
Certification.
> >> When
> >> > > the semiconductor junctions are unbiased and cool off when power is
> >> lost,
> >> > > and then power is restored the result will be a different 10 volts
> >> than
> >> > > before the power failure. My experience is that after all of the
> >> years
> >> > > that
> >> > > these units have been powered up, this won't happen and when power
is
> >> lost
> >> > > and then restored, even months later, the 732A will come back to
> >> almost
> >> > > exactly the same 10 volts as when they lost power, usually with in
0.2
> >> PPM
> >> > > after 24 hours of "warm up".
> >> > >
> >> > > What type of hookup leads are you using when measuring the 1
volt
> >> > > output? What is the PLC set to? I always use 100 PLC to measure
> >> this.
> >> If
> >> > > you don't have "low thermal" connection leads you can experience uV
> >> changes
> >> > > for a minute or more after plugging in the leads due to the
"thermals"
> >> > > generated because of the difference in temperature between the
banana
> >> jacks
> >> > > on the 732A and the banana plugs of the leads. I have found that
even
> >> just
> >> > > plugging in the lead will generate a thermal difference because of
> >> > > difference of temps and some heating due to the physical act of
just
> >> > > inserting the plug because of friction between the jack and plug
(my
> >> theory
> >> > > at any rate). You have to allow at least a minute or more before
> >> being
> >> > > able
> >> > > to make a measurement after plugging in the leads. I just measured
> >> the
> >> > > variation of the 1 volt output of my 732A and using my 3458A and I
> >> got a
> >> > > total difference of 0.159 uV over 40 measurements using 100 PLC on
> >> the
> >> 1
> >> > > volt range of the 3458A. Using the MATH function and all of the
data
> >> you
> >> > > can collect. That was after waiting for several minutes after
> >> plugging
> >> in
> >> > > the leads.
> >> > >
> >> > > I hope all of this helps.
> >> > >
> >> > > Bill
> >> > >
> >> > >
> >> > >
> >> > > ----- Original Message -----
> >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> >> > > To: "Discussion of precise voltage measurement"
<volt-nuts@febo.com>
> >> > > Sent: Friday, August 22, 2014 9:03 PM
> >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> >> > >
> >> > >
> >> > > > Todd,
> >> > > >
> >> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> >> that
> >> I
> >> > > > keep topped off and they have very low current draw (~2 to 3 mA
at
> >> 13.5
> >> > > > VDC) when charged and at their float voltage, so I am pretty sure
> >> they
> >> > > are
> >> > > > in good condition. I will look at getting those in the units
after
> >> I
> >> > > > ascertain the condition of the 732.
> >> > > >
> >> > > > So now I have a what appears to be a functioning 3458A and a 732A
> >> but
> >> > > they
> >> > > > slightly disagree. I am like the man with two watches that
disagree
> >> on
> >> > > the
> >> > > > time - which is correct? For the moment, i am only concerned
with
> >> > > > stability. The need for absolute accuracy will come later.
> >> > > >
> >> > > > Randy
> >> > > >
> >> > > >
> >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
<tmicallef@gmail.com
> >> >
> >> > > wrote:
> >> > > >
> >> > > > > Randy,
> >> > > > >
> >> > > > > You have two possible choices. It can be configured with 4 x 6v
> >> 4Ah
> >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
owner
> >> has
> >> > > > > modified the battery pack already. A couple of mine needed a
> >> nibbler
> >> > > tool
> >> > > > > to remove enough of the aluminum cover that fits over the tops
of
> >> the
> >> > > > > batteries. The original cover will short out to the battery
tabs
> >> > > regardless
> >> > > > > of the battery configuration if this is not done.
> >> > > > >
> >> > > > > You can find larger capacity batteries that will give you
slightly
> >> more
> >> > > > > battery life. I lost a couple sets of mail-order batteries
after a
> >> few
> >> > > > > extended outages. I would recommend going with locally bought
> >> batteries
> >> > > > > instead of the cheaper mail order. My local Batteries Plus will
> >> > > typically
> >> > > > > have some warranty if I remember correctly. Moving forward I
will
> >> only
> >> > > use
> >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> >> equalize
> >> > > them
> >> > > > > before putting them in the 732A. I think the cheap batteries
did
> >> not
> >> > > > > discharge equally, and would not recover when power was
applied.
> >> > > > >
> >> > > > > Inspect the back plane for damaged traces and look at the
> >> capacitors. I
> >> > > had
> >> > > > > a few that looked questionable. So far, I have replaced all the
> >> big
> >> > > caps
> >> > > on
> >> > > > > the pre-regulator and regulator boards. My feeling is that once
> >> these
> >> > > go
> >> > > > > online, they should run as long as possible between repairs.
> >> > > > >
> >> > > > > The battery charger circuit may need adjusting. I tweaked mine
and
> >> it
> >> > > > > seemed to work fine.
> >> > > > >
> >> > > > > Todd
> >> > > > >
> >> > > > >
> >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> >> > > randyevans2688@gmail.com>
> >> > > > > wrote:
> >> > > > >
> >> > > > > > I received my Fluke 732A today. Just powered it up but it
needs
> >> new
> >> > > > > > batteries. Any suggestions for sources (I haven't opened up
the
> >> unit
> >> > > > > yet -
> >> > > > > > I want to make sure it works before doing that). Also
received
> >> the
> >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
ran
> >> data
> >> > > > > dumper
> >> > > > > > program to get the CAL data from my 3458A. Should be a busy
> >> weekend.
> >> > > > > >
> >> > > > > > Randy
> >> > > > > > _______________________________________________
> >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > > > To unsubscribe, go to
> >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > > > and follow the instructions there.
> >> > > > > >
> >> > > > > _______________________________________________
> >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > > To unsubscribe, go to
> >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > > and follow the instructions there.
> >> > > > >
> >> > > > _______________________________________________
> >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > > To unsubscribe, go to
> >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > > and follow the instructions there.
> >> > >
> >> > > _______________________________________________
> >> > > volt-nuts mailing list -- volt-nuts@febo.com
> >> > > To unsubscribe, go to
> >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > > and follow the instructions there.
> >> > >
> >> > _______________________________________________
> >> > volt-nuts mailing list -- volt-nuts@febo.com
> >> > To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> > and follow the instructions there.
> >>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >>
> >
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
MS
Mike S
Tue, Aug 26, 2014 3:05 PM
After some more research, I think I've answered some of my own questions -
Tellurium copper is used for binding posts, not because it has any
special thermal or EMF mojo, but because it machines much better than
pure copper. And, I suppose, because it sounds like it's extra special.
The Seebeck coefficients (uV/C, relative to Cu) of some relevant materials:
Cu 0.0
Ag .2
Au .5
Yellow brass 1.5
Phosphor bronze 2.0
63/37 solder 3.0
Sn 3.1
Stainless steel 3.1
Beryllium copper 5.0
Fe -12.3
Ni 22.3
Te -49.25
Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
I'd expect that there might be some coefficient between Cu and CuTe
(0.5% Te), but I could find no reference. The relatively large number
for CuBe is interesting, since that's a common material for banana plug
springs, where one might expect the greatest temperature differential to
occur in such a connection (between the thermal masses of the binding
post/jack and the bulk of the banana plug). Heat has to flow a
considerable distance through the springs, very much more than when it
flows through a surface plating.
The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
contacts. It seems there might be an improvement to be had by using the
older style pin plugs, where a solid pin was partially sliced into 4
sections which were then spread apart a bit to create tension. That
could eliminate relatively large thermocouples at a thermal gradient,
and might also be expected to have less thermal resistance, allowing the
connection to settle quicker.
But maybe not - I'm still not clear on how plated conductors behave in
this situation. For a high impedance voltage measurement where almost no
current flows, the gold plating may carry the signal, so there is no
real thermocouple (or more correctly, it's entirely contained within the
connector). But if that's the case, why fool around with special copper
connectors when common brass ones would be easier/cheaper? For current
or resistance, the signal would also flow through the base metal, so
does this have an effect (especially for tinned copper test leads, where
there may be a larger temperature difference between the ends???
Nickle is avoided as a contact material largely because it is subject to
fretting corrosion. Tests done by AMP
(http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show that
a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a short
time due to this, while Ag and Au plated contacts exhibit negligible
changes.
Cu (with Be for better machining) seems to be used as the base material
for jacks/plugs to get thermal EMF cancellation to the wiring on both
sides (i.e. use copper everywhere except where there is a minimal
thermal gradient, like platings).
--
After some more research, I think I've answered some of my own questions -
Tellurium copper is used for binding posts, not because it has any
special thermal or EMF mojo, but because it machines much better than
pure copper. And, I suppose, because it sounds like it's extra special.
The Seebeck coefficients (uV/C, relative to Cu) of some relevant materials:
Cu 0.0
Ag .2
Au .5
Yellow brass 1.5
Phosphor bronze 2.0
63/37 solder 3.0
Sn 3.1
Stainless steel 3.1
Beryllium copper 5.0
Fe -12.3
Ni 22.3
Te -49.25
Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
I'd expect that there might be some coefficient between Cu and CuTe
(0.5% Te), but I could find no reference. The relatively large number
for CuBe is interesting, since that's a common material for banana plug
springs, where one might expect the greatest temperature differential to
occur in such a connection (between the thermal masses of the binding
post/jack and the bulk of the banana plug). Heat has to flow a
considerable distance through the springs, very much more than when it
flows through a surface plating.
The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
contacts. It seems there might be an improvement to be had by using the
older style pin plugs, where a solid pin was partially sliced into 4
sections which were then spread apart a bit to create tension. That
could eliminate relatively large thermocouples at a thermal gradient,
and might also be expected to have less thermal resistance, allowing the
connection to settle quicker.
But maybe not - I'm still not clear on how plated conductors behave in
this situation. For a high impedance voltage measurement where almost no
current flows, the gold plating may carry the signal, so there is no
real thermocouple (or more correctly, it's entirely contained within the
connector). But if that's the case, why fool around with special copper
connectors when common brass ones would be easier/cheaper? For current
or resistance, the signal would also flow through the base metal, so
does this have an effect (especially for tinned copper test leads, where
there may be a larger temperature difference between the ends???
Nickle is avoided as a contact material largely because it is subject to
fretting corrosion. Tests done by AMP
(http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show that
a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a short
time due to this, while Ag and Au plated contacts exhibit negligible
changes.
Cu (with Be for better machining) seems to be used as the base material
for jacks/plugs to get thermal EMF cancellation to the wiring on both
sides (i.e. use copper everywhere except where there is a minimal
thermal gradient, like platings).
--
A
acbern@gmx.de
Tue, Aug 26, 2014 4:56 PM
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Dave M
acbern@gmx.de wrote:
fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be <10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.
adrian
Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer. Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer. At that point,
the AC voltage source is equal to that of the DC voltage source.
Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard. They aren't cheap, and you
have to have a converter for each range of voltages that you need to
measure. The thermal converters used with this type of transfer
standard isn't great (50 MHz or so typical), but their accuracy far
surpasses that of the thermistor type sensors.
There are other brands and models of thermal transfer standards, but
I have a Fluke model 540 and a few thermal converters. That's why I
referred you to the manual for it.
Cheers,
Dave M
pa4tim@gmail.com wrote:
Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two
AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
ago.
My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for
1V, 10V and 100V (and use several resistors/heaters.)
Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that) I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.
Fred, pa4tim
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.
Adrian
> Gesendet: Montag, 25. August 2014 um 19:32 Uhr
> Von: "Dave M" <dgminala@mediacombb.net>
> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Betreff: Re: [volt-nuts] AC calibration
>
> Adrian,
> Do you have a link or title for the NIST paper that you mentioned?
>
> Dave M
>
>
> acbern@gmx.de wrote:
> > fred,
> > generally you raise a good point, I had the same issue of calibrating
> > an ac voltage to a high level of accuracy. you need this e.g. to
> > validate the self.cal of a 3458a or other precison stuff like the
> > 8506a0.
> >
> > what i would recommend to do if you want to keep costs down is:
> > in a nutshell, get a thermal converter in the lowest range you need
> > and a second one on range above. build a set of resistor range
> > extenders (rf type with appropriate connectors and housings) to
> > expand the range to where you need to be max. get one of the thermal
> > converter calibrated (the higher one usually, and you need to havr
> > good cal lab, should be <10ppm accuracy) and use it to calibrate the
> > rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
> > thermal converters! at higher frequencies, due to reflections and
> > stray capacitance/inductance influences, the accuracy decreases. the
> > resistor range extenders though, if build up correctly, only have a
> > few ppm impact (there is a paper from nist on that, but this is only
> > typical). you can calibrate all converters to the one you got
> > externally calibrated. do some research in the web, when you do the
> > calibration, you need to determine the so-called constant N. then do
> > an ac, dc+, ac, dc-, ac measurement between the the two and establish
> > the deviation, also establish the error propagation. the end result
> > will be a set of highly precise (low inaccuracies9 thermal converters
> > good enough to calibrate a 3458a an better devices. if you want to
> > spend the money, you could also buy a set of converters/range
> > resistors (with/without a 540), that typically is a few k altogether,
> > while a single device sometimes is available for below 100 bucks. you
> > need to have a stable 7.5 digit nanovoltmeter though for the
> > measurements of the tvcs (34420a or 2182 typically ) and precision
> > (stable) dc and ac sources. but in the end, all you need is a single
> > calibrated thermal converter.
> >
> > adrian
> >
> >
> >
> >> Gesendet: Montag, 25. August 2014 um 18:38 Uhr
> >> Von: "Dave M" <dgminala@mediacombb.net>
> >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> >> Betreff: Re: [volt-nuts] AC calibration
> >>
> >> Well, you sort of answered your own question. The equipment is
> >> called a Thermal Transfer Standard, but instead of thermistors, it
> >> uses a thermocouple. Look at the manual for the Fluke 540B
> >> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
> >> it's done. Basically, the AC source is input into the transfer
> >> standard, and the standard's internal reference voltage is adjusted
> >> for a null on the galvanometer. Leaving the reference voltage
> >> setting alone, a DC voltage is input into the unit, and the DC
> >> source is adjusted for a null on the galvanometer. At that point,
> >> the AC voltage source is equal to that of the DC voltage source.
> >>
> >> Ther are thermocouple-type thermal converters used for RF voltage
> >> measurements with the transfer standard. They aren't cheap, and you
> >> have to have a converter for each range of voltages that you need to
> >> measure. The thermal converters used with this type of transfer
> >> standard isn't great (50 MHz or so typical), but their accuracy far
> >> surpasses that of the thermistor type sensors.
> >>
> >> There are other brands and models of thermal transfer standards, but
> >> I have a Fluke model 540 and a few thermal converters. That's why I
> >> referred you to the manual for it.
> >>
> >> Cheers,
> >> Dave M
> >>
> >>
> >> pa4tim@gmail.com wrote:
> >>> Is there a way to link an AC voltage to a DC source for compare. I
> >>> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
> >>> against standardcells. But for AC I can not do that. I have two
> >>> AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
> >>> ago.
> >>>
> >>> My idea is in theory simple. It is based on the thermal converters
> >>> used in RF powermeters. Two resistors, two high resolution
> >>> temperature meters. AC on the first en DC on the second. If both are
> >>> the same temperature the AC voltage is the same as the DC voltage.
> >>> But I'm sure some people here have done this in the past. I would
> >>> like to use it for 50 to 100 kHz (or less) and something like for
> >>> 1V, 10V and 100V (and use several resistors/heaters.)
> >>>
> >>> Or mabey there is an other way to convert AC (for RF it can be done
> >>> with lightbubs but I never tryed that) I do not mind if it is slow
> >>> etc, I like this sort of experiments. You can learn a lot from it.
> >>>
> >>> Fred, pa4tim
> >>
> >>
> >> _______________________________________________
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
> >> the instructions there.
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> When the people fear the government, there is tyranny. When the
> government fears the people, there is liberty -- Thomas Jefferson
>
>
> Dave M
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
PK
Poul-Henning Kamp
Tue, Aug 26, 2014 7:42 PM
In message <trinity-9076c25b-a0a9-4b28-b29a-20f9e06039a3-1409072175101@3capp-gm
x-bs56>, acbern@gmx.de writes:
Tellurium/copper is used because it machines much better than pure
copper, (20% -> 90%) without ruining the conductivity too much.
Most other metals which can improve its machinability has big
negative impacts on the conductivity of copper.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
--------
In message <trinity-9076c25b-a0a9-4b28-b29a-20f9e06039a3-1409072175101@3capp-gm
x-bs56>, acbern@gmx.de writes:
Tellurium/copper is used because it machines much better than pure
copper, (20% -> 90%) without ruining the conductivity too much.
Most other metals which can improve its machinability has big
negative impacts on the conductivity of copper.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
RE
Randy Evans
Tue, Aug 26, 2014 9:16 PM
Bill,
I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.
Randy
On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3637@att.net wrote:
Not that I know of. Just wait and when you don't see a "SMPL" on the
display it is done. But then with 1000 PLC that is around 16.66 seconds
per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes. Not from the front panel
at any rate.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Is there any way to tell when the function key routine is complete? In
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete. In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long
before it's complete, but i have to guess when it's done.
Thanks,
Randy
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:
Bill,
I am trying to figure out the MATH function without much success. I
the sequence you said (I looked up the instructions to understand what
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
it shows up on the display when I input BLUE F1. I hit ENTER and it
the 40 measurements and the MATH symbol shows on the display during the
measurements. After the SMPL symbol no longer blinks I hit MATH 2 and
get a MATH ERR symbol on the display. I tried it a couple of times and
same result so I am doing something wrong. Is there a better source
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual
Randy:
The MATH function is accessible from the keypad. I don't have an
interface right now that works. You can also program the numeric
keys to have preprogrammed functions. DEFKEY
I have made my own "low thermal" measurement leads from Pomona
banana plugs and Belden #9272 wire. Why 9272, because it was handy at
time. It is tin plated copper, shielded twisted pair 20 ga. I have
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it. I simply cannot find what I want so I
build my own cable. I have done something like this before and it
fine. When I get a "round toit".
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
used
in the past to verify my homemade "low thermal" leads as described
Frankly I cannot see any difference between using the 1756 cables and
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below
ppm
at 10 volts.
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
adjustable you may be seeing a misadjusted 1 volt from the 732A. As
as
the instability of the readings it is hard to determine which is
the
problem. I have programed (DEFKEY) a numeric keypad key #1 with the
following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
set
the MATH to "Statistics" (store high reading/low reading/ and mean of
readings) in the registers, the number of readings to "40", the
"hold" (which keeps the meter from triggering until I press "ENTER"
then
trigger the sequence of 40 readings when I push the "ENTER" button.
can
do all of this manually from the keypads but since I use this sequence
lot
I have preprogrammed it. This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various
statistic registers, using the menu, by entering "MATH" and then a 2
low, a 4 for mean, and 13 for high. Of course you could do all of
through the IEEE also. The 3458A has a very rich set of measurement
commands. I am still learning all of them. It depends upon what I am
trying to accomplish.
Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could
to the varying readings you are measuring. I think I would put a
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved. When I do this I
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts. This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect
readings. So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output
got
a variation of 0.159 uVolts using the same 40 reading method above. I
would
use this to determine where your problem might exist. Just having the
meter
input shorted will point you in the right direction. Meter, cables or
732A.
Sorry for the long dissertation. Friends get mad at me for being
detailed sometimes.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received
Bill,
I have convinced myself that the problem I an seeing is due to
If I move the cables (with gold-plated banana plugs) using a small
rather than letting my hand touch the plugs, it is much more stable.
then hold the banana plug with my hand after the reading has
the reading drifts rapidly upward. I am trying to check the
the reading but I haven't figured out the MATH function yet. I
is a programmed function using GPIB only?
The stability I am seeing by manually recording readings (using NLPC
and 1000) is much greater than what you are measuring on your
sure how to ascertain if it's the 3458A or the 732. The value of
readings are very different between the two - the 3458 reads about
high on the 10 V output and about 12 uV low on the 1V output.
differences (this is after an ACAL). I need to find some better
make sure the errors are not due to thermals again.
Randy
On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net
Randy:
I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
others
that are in this size and package. Order from one of the usual
distributors like Allied, Mouser, Digikey. This is a very common
as
it is used in a lot of "EXIT" signs so they are lighted when the
out. I don't see how the 12V 7AH will fit as they are too large.
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
but the terminals are in the wrong place so you will have to
the aluminum plate that holds them in the 732A battery pack. You
be
careful if you use the 12v 5AH as you will have 4 extra battery
leads to deal with and connect correctly. I would stick with the
New batteries will last around 12 to 14 hours before the "CAL"
out when AC power is not applied. So shipping to Cal Lab can be a
if it is a distance away, or you have to use a shipper like UPS or
and
you ship the night before and then use their "Morning delivery"
Lab is expecting your 732A. Same on the way back to you. Of
could always strap another battery on the 732A and hook it up to
power" plug to last longer. I have seen it done. The issue is to
Cal Lab to charge the extra battery before they ship the 732A back
When you remove and work on the battery pack always have the
plugged in. The "CAL" led will stay on because the 18.6 v
is working.
The "CAL" light is to indicate that power has not been lost to
Reference Amp or other associated circuits. When the raw supply
voltage drops below about 21 volts the "CAL" light will go out.
voltage the heater circuits will not work correctly and the 18.6
regulated supply will not regulate. The requirement is that the
Amp be kept "alive" at all times to maintain the output voltage
measured at the time of the most recent Calibration or
the semiconductor junctions are unbiased and cool off when power
and then power is restored the result will be a different 10 volts
before the power failure. My experience is that after all of the
that
these units have been powered up, this won't happen and when power
and then restored, even months later, the 732A will come back to
exactly the same 10 volts as when they lost power, usually with in
after 24 hours of "warm up".
What type of hookup leads are you using when measuring the 1
output? What is the PLC set to? I always use 100 PLC to measure
you don't have "low thermal" connection leads you can experience
for a minute or more after plugging in the leads due to the
generated because of the difference in temperature between the
on the 732A and the banana plugs of the leads. I have found that
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of
inserting the plug because of friction between the jack and plug
at any rate). You have to allow at least a minute or more before
able
to make a measurement after plugging in the leads. I just
variation of the 1 volt output of my 732A and using my 3458A and I
total difference of 0.159 uV over 40 measurements using 100 PLC
volt range of the 3458A. Using the MATH function and all of the
can collect. That was after waiting for several minutes after
the leads.
I hope all of this helps.
Bill
----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received
Todd,
Thanks for the info. I have several Panasonic 12V 7 AH
keep topped off and they have very low current draw (~2 to 3 mA
VDC) when charged and at their float voltage, so I am pretty
in good condition. I will look at getting those in the units
ascertain the condition of the 732.
So now I have a what appears to be a functioning 3458A and a
slightly disagree. I am like the man with two watches that
time - which is correct? For the moment, i am only concerned
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
Randy,
You have two possible choices. It can be configured with 4 x
batteries or 2 x 12v 7Ah batteries. Hopefully the previous
modified the battery pack already. A couple of mine needed a
to remove enough of the aluminum cover that fits over the tops
batteries. The original cover will short out to the battery
of the battery configuration if this is not done.
You can find larger capacity batteries that will give you
battery life. I lost a couple sets of mail-order batteries
extended outages. I would recommend going with locally bought
instead of the cheaper mail order. My local Batteries Plus
have some warranty if I remember correctly. Moving forward I
2 12v batteries and pre-charge them on a battery charger to
before putting them in the 732A. I think the cheap batteries
discharge equally, and would not recover when power was
Inspect the back plane for damaged traces and look at the
a few that looked questionable. So far, I have replaced all
the pre-regulator and regulator boards. My feeling is that
online, they should run as long as possible between repairs.
The battery charger circuit may need adjusting. I tweaked mine
seemed to work fine.
Todd
On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
I received my Fluke 732A today. Just powered it up but it
batteries. Any suggestions for sources (I haven't opened up
I want to make sure it works before doing that). Also
ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
program to get the CAL data from my 3458A. Should be a busy
and follow the instructions there.
and follow the instructions there.
and follow the instructions there.
Bill,
I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.
Randy
On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold <wpgold3637@att.net> wrote:
> Not that I know of. Just wait and when you don't see a "SMPL" on the
> display it is done. But then with 1000 PLC that is around 16.66 seconds
> per
> reading times 100 readings is somewhere around 28 minutes and there is
> probably some overhead time so around 30 minutes. Not from the front panel
> at any rate.
>
> Bill
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Monday, August 25, 2014 9:23 PM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Is there any way to tell when the function key routine is complete? In
> the
> > case of taking multiple readings using the DEFKEY and MATH function, I
> > don't see any indication when the routine is complete. In one particular
> > case, I am taking a 100 readings with NLPC set for 1000 so its a long
> while
> > before it's complete, but i have to guess when it's done.
> >
> > Thanks,
> >
> > Randy
> >
> >
> > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com>
> > wrote:
> >
> > > Bill,
> > >
> > > I am trying to figure out the MATH function without much success. I
> input
> > > the sequence you said (I looked up the instructions to understand what
> you
> > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
> and
> > > it shows up on the display when I input BLUE F1. I hit ENTER and it
> takes
> > > the 40 measurements and the MATH symbol shows on the display during the
> > > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and
> I
> > > get a MATH ERR symbol on the display. I tried it a couple of times and
> the
> > > same result so I am doing something wrong. Is there a better source
> for
> > > explaining how to do front panel masurements than the User Guide, which
> > > seems oriented at programming automatic rather than manual
> measurements.
> > >
> > > Randy
> > >
> > >
> > >
> > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote:
> > >
> > >> Randy:
> > >>
> > >> The MATH function is accessible from the keypad. I don't have an
> IEEE
> > >> interface right now that works. You can also program the numeric
> keypad
> > >> keys to have preprogrammed functions. DEFKEY
> > >>
> > >> I have made my own "low thermal" measurement leads from Pomona
> #4892
> > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at
> the
> > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have
> plans
> > >> to do custom cables with 16 ga. bare copper wire that I will twist and
> > >> then
> > >> put a braided shield over it. I simply cannot find what I want so I
> will
> > >> build my own cable. I have done something like this before and it
> worked
> > >> fine. When I get a "round toit".
> > >>
> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
> have
> > >> used
> > >> in the past to verify my homemade "low thermal" leads as described
> above.
> > >> Frankly I cannot see any difference between using the 1756 cables and
> my
> > >> homemade cables once I give them a few minutes for the thermals to go
> > >> away.
> > >> As far as I can tell and measure the differences, if any, are below
> 0.1
> > >> ppm
> > >> at 10 volts.
> > >>
> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
> all
> > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As
> far
> > >> as
> > >> the instability of the readings it is hard to determine which is
> causing
> > >> the
> > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does
> is
> > >> set
> > >> the MATH to "Statistics" (store high reading/low reading/ and mean of
> the
> > >> readings) in the registers, the number of readings to "40", the
> trigger
> to
> > >> "hold" (which keeps the meter from triggering until I press "ENTER"
> and
> > >> then
> > >> trigger the sequence of 40 readings when I push the "ENTER" button.
> You
> > >> can
> > >> do all of this manually from the keypads but since I use this sequence
> a
> > >> lot
> > >> I have preprogrammed it. This is after I set digits to "8" and PLC to
> > >> 100.
> > >> Once those 40 readings are finished then you can access the various
> MATH
> > >> statistic registers, using the menu, by entering "MATH" and then a 2
> for
> > >> low, a 4 for mean, and 13 for high. Of course you could do all of
> this
> > >> through the IEEE also. The 3458A has a very rich set of measurement
> > >> commands. I am still learning all of them. It depends upon what I am
> > >> trying to accomplish.
> > >>
> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from
> > >> resistive dividers from the 10 volt, I don't see how they could
> contribute
> > >> to the varying readings you are measuring. I think I would put a
> short
> on
> > >> the input of the 3458A and manually set the range to 1 volt and then
> > >> observe
> > >> the variations that way without the 732A involved. When I do this I
> see a
> > >> variation from low reading to high reading of 0.125 uVolts and then
> > >> another
> > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low
> > >> side
> > >> of the measurment terminals, GUARD connected doesn't seem to affect
> the
> > >> readings. So that is the base noise of the 3458A without the 732A,
> > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output
> I
> > >> got
> > >> a variation of 0.159 uVolts using the same 40 reading method above. I
> > >> would
> > >> use this to determine where your problem might exist. Just having the
> > >> meter
> > >> input shorted will point you in the right direction. Meter, cables or
> > >> 732A.
> > >>
> > >> Sorry for the long dissertation. Friends get mad at me for being
> so
> > >> detailed sometimes.
> > >>
> > >> Bill
> > >>
> > >> ----- Original Message -----
> > >> From: "Randy Evans" <randyevans2688@gmail.com>
> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> > >> Sent: Sunday, August 24, 2014 7:22 AM
> > >> Subject: Re: [volt-nuts] 732A and Prologix received
> > >>
> > >>
> > >> > Bill,
> > >> >
> > >> > I have convinced myself that the problem I an seeing is due to
> thermals.
> > >> > If I move the cables (with gold-plated banana plugs) using a small
> > >> towel
> > >> > rather than letting my hand touch the plugs, it is much more stable.
> > >> If I
> > >> > then hold the banana plug with my hand after the reading has
> stabilized,
> > >> > the reading drifts rapidly upward. I am trying to check the
> stability
> > >> of
> > >> > the reading but I haven't figured out the MATH function yet. I
> assume
> > >> this
> > >> > is a programmed function using GPIB only?
> > >> >
> > >> > The stability I am seeing by manually recording readings (using NLPC
> of
> > >> 100
> > >> > and 1000) is much greater than what you are measuring on your
> system.
> > >> Not
> > >> > sure how to ascertain if it's the 3458A or the 732. The value of
> the
> > >> > readings are very different between the two - the 3458 reads about
> 50
> uV
> > >> > high on the 10 V output and about 12 uV low on the 1V output.
> Rather
> > >> large
> > >> > differences (this is after an ACAL). I need to find some better
> cables
> > >> to
> > >> > make sure the errors are not due to thermals again.
> > >> >
> > >> > Randy
> > >> >
> > >> >
> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net>
> wrote:
> > >> >
> > >> > > Randy:
> > >> > >
> > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will
> fit
> > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C
> and
> > >> > > others
> > >> > > that are in this size and package. Order from one of the usual
> > >> electronics
> > >> > > distributors like Allied, Mouser, Digikey. This is a very common
> > >> battery
> > >> > > as
> > >> > > it is used in a lot of "EXIT" signs so they are lighted when the
> power
> > >> goes
> > >> > > out. I don't see how the 12V 7AH will fit as they are too large.
> I
> > >> guess
> > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6
> volt
> > >> 4
> > >> AH
> > >> > > but the terminals are in the wrong place so you will have to
> "nibble"
> > >> out
> > >> > > the aluminum plate that holds them in the 732A battery pack. You
> have
> > >> to
> > >> > > be
> > >> > > careful if you use the 12v 5AH as you will have 4 extra battery
> > >> connection
> > >> > > leads to deal with and connect correctly. I would stick with the
> 6V
> > >> 4AH.
> > >> > > New batteries will last around 12 to 14 hours before the "CAL"
> light
> > >> goes
> > >> > > out when AC power is not applied. So shipping to Cal Lab can be a
> > >> problem
> > >> > > if it is a distance away, or you have to use a shipper like UPS or
> > >> FEDEX
> > >> > > and
> > >> > > you ship the night before and then use their "Morning delivery"
> and
> > >> the
> > >> Cal
> > >> > > Lab is expecting your 732A. Same on the way back to you. Of
> course
> > >> you
> > >> > > could always strap another battery on the 732A and hook it up to
> the
> > >> "ext
> > >> > > power" plug to last longer. I have seen it done. The issue is to
> get
> > >> the
> > >> > > Cal Lab to charge the extra battery before they ship the 732A back
> to
> > >> you.
> > >> > >
> > >> > > When you remove and work on the battery pack always have the
> AC
> > >> power
> > >> > > plugged in. The "CAL" led will stay on because the 18.6 v
> regulated
> > >> supply
> > >> > > is working.
> > >> > >
> > >> > > The "CAL" light is to indicate that power has not been lost to
> the
> > >> > > Reference Amp or other associated circuits. When the raw supply
> > >> (battery)
> > >> > > voltage drops below about 21 volts the "CAL" light will go out.
> Below
> > >> that
> > >> > > voltage the heater circuits will not work correctly and the 18.6
> volt
> > >> > > regulated supply will not regulate. The requirement is that the
> > >> Reference
> > >> > > Amp be kept "alive" at all times to maintain the output voltage
> that
> > >> was
> > >> > > measured at the time of the most recent Calibration or
> Certification.
> > >> When
> > >> > > the semiconductor junctions are unbiased and cool off when power
> is
> > >> lost,
> > >> > > and then power is restored the result will be a different 10 volts
> > >> than
> > >> > > before the power failure. My experience is that after all of the
> > >> years
> > >> > > that
> > >> > > these units have been powered up, this won't happen and when power
> is
> > >> lost
> > >> > > and then restored, even months later, the 732A will come back to
> > >> almost
> > >> > > exactly the same 10 volts as when they lost power, usually with in
> 0.2
> > >> PPM
> > >> > > after 24 hours of "warm up".
> > >> > >
> > >> > > What type of hookup leads are you using when measuring the 1
> volt
> > >> > > output? What is the PLC set to? I always use 100 PLC to measure
> > >> this.
> > >> If
> > >> > > you don't have "low thermal" connection leads you can experience
> uV
> > >> changes
> > >> > > for a minute or more after plugging in the leads due to the
> "thermals"
> > >> > > generated because of the difference in temperature between the
> banana
> > >> jacks
> > >> > > on the 732A and the banana plugs of the leads. I have found that
> even
> > >> just
> > >> > > plugging in the lead will generate a thermal difference because of
> > >> > > difference of temps and some heating due to the physical act of
> just
> > >> > > inserting the plug because of friction between the jack and plug
> (my
> > >> theory
> > >> > > at any rate). You have to allow at least a minute or more before
> > >> being
> > >> > > able
> > >> > > to make a measurement after plugging in the leads. I just
> measured
> > >> the
> > >> > > variation of the 1 volt output of my 732A and using my 3458A and I
> > >> got a
> > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC
> on
> > >> the
> > >> 1
> > >> > > volt range of the 3458A. Using the MATH function and all of the
> data
> > >> you
> > >> > > can collect. That was after waiting for several minutes after
> > >> plugging
> > >> in
> > >> > > the leads.
> > >> > >
> > >> > > I hope all of this helps.
> > >> > >
> > >> > > Bill
> > >> > >
> > >> > >
> > >> > >
> > >> > > ----- Original Message -----
> > >> > > From: "Randy Evans" <randyevans2688@gmail.com>
> > >> > > To: "Discussion of precise voltage measurement"
> <volt-nuts@febo.com>
> > >> > > Sent: Friday, August 22, 2014 9:03 PM
> > >> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >> > >
> > >> > >
> > >> > > > Todd,
> > >> > > >
> > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH
> batteries
> > >> that
> > >> I
> > >> > > > keep topped off and they have very low current draw (~2 to 3 mA
> at
> > >> 13.5
> > >> > > > VDC) when charged and at their float voltage, so I am pretty
> sure
> > >> they
> > >> > > are
> > >> > > > in good condition. I will look at getting those in the units
> after
> > >> I
> > >> > > > ascertain the condition of the 732.
> > >> > > >
> > >> > > > So now I have a what appears to be a functioning 3458A and a
> 732A
> > >> but
> > >> > > they
> > >> > > > slightly disagree. I am like the man with two watches that
> disagree
> > >> on
> > >> > > the
> > >> > > > time - which is correct? For the moment, i am only concerned
> with
> > >> > > > stability. The need for absolute accuracy will come later.
> > >> > > >
> > >> > > > Randy
> > >> > > >
> > >> > > >
> > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef
> <tmicallef@gmail.com
> > >> >
> > >> > > wrote:
> > >> > > >
> > >> > > > > Randy,
> > >> > > > >
> > >> > > > > You have two possible choices. It can be configured with 4 x
> 6v
> > >> 4Ah
> > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous
> owner
> > >> has
> > >> > > > > modified the battery pack already. A couple of mine needed a
> > >> nibbler
> > >> > > tool
> > >> > > > > to remove enough of the aluminum cover that fits over the tops
> of
> > >> the
> > >> > > > > batteries. The original cover will short out to the battery
> tabs
> > >> > > regardless
> > >> > > > > of the battery configuration if this is not done.
> > >> > > > >
> > >> > > > > You can find larger capacity batteries that will give you
> slightly
> > >> more
> > >> > > > > battery life. I lost a couple sets of mail-order batteries
> after a
> > >> few
> > >> > > > > extended outages. I would recommend going with locally bought
> > >> batteries
> > >> > > > > instead of the cheaper mail order. My local Batteries Plus
> will
> > >> > > typically
> > >> > > > > have some warranty if I remember correctly. Moving forward I
> will
> > >> only
> > >> > > use
> > >> > > > > 2 12v batteries and pre-charge them on a battery charger to
> > >> equalize
> > >> > > them
> > >> > > > > before putting them in the 732A. I think the cheap batteries
> did
> > >> not
> > >> > > > > discharge equally, and would not recover when power was
> applied.
> > >> > > > >
> > >> > > > > Inspect the back plane for damaged traces and look at the
> > >> capacitors. I
> > >> > > had
> > >> > > > > a few that looked questionable. So far, I have replaced all
> the
> > >> big
> > >> > > caps
> > >> > > on
> > >> > > > > the pre-regulator and regulator boards. My feeling is that
> once
> > >> these
> > >> > > go
> > >> > > > > online, they should run as long as possible between repairs.
> > >> > > > >
> > >> > > > > The battery charger circuit may need adjusting. I tweaked mine
> and
> > >> it
> > >> > > > > seemed to work fine.
> > >> > > > >
> > >> > > > > Todd
> > >> > > > >
> > >> > > > >
> > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > >> > > randyevans2688@gmail.com>
> > >> > > > > wrote:
> > >> > > > >
> > >> > > > > > I received my Fluke 732A today. Just powered it up but it
> needs
> > >> new
> > >> > > > > > batteries. Any suggestions for sources (I haven't opened up
> the
> > >> unit
> > >> > > > > yet -
> > >> > > > > > I want to make sure it works before doing that). Also
> received
> > >> the
> > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL
> ran
> > >> data
> > >> > > > > dumper
> > >> > > > > > program to get the CAL data from my 3458A. Should be a busy
> > >> weekend.
> > >> > > > > >
> > >> > > > > > Randy
> > >> > > > > > _______________________________________________
> > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > > > To unsubscribe, go to
> > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > > > and follow the instructions there.
> > >> > > > > >
> > >> > > > > _______________________________________________
> > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > > To unsubscribe, go to
> > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > > and follow the instructions there.
> > >> > > > >
> > >> > > > _______________________________________________
> > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > > To unsubscribe, go to
> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > > and follow the instructions there.
> > >> > >
> > >> > > _______________________________________________
> > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > > To unsubscribe, go to
> > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > > and follow the instructions there.
> > >> > >
> > >> > _______________________________________________
> > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > >> > To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> > and follow the instructions there.
> > >>
> > >> _______________________________________________
> > >> volt-nuts mailing list -- volt-nuts@febo.com
> > >> To unsubscribe, go to
> > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > >> and follow the instructions there.
> > >>
> > >
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
DM
Dave M
Wed, Aug 27, 2014 12:01 AM
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC
Voltage Measurements by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Great!!@! Many Thanks!!
Dave M
acbern@gmx.de wrote:
> Dave,
>
> the title is:
> Thermal Voltage Converters and Comparator for Very Accurate AC
> Voltage Measurements by E.S.Williams.
>
> Adrian
>
>
>> Gesendet: Montag, 25. August 2014 um 19:32 Uhr
>> Von: "Dave M" <dgminala@mediacombb.net>
>> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Betreff: Re: [volt-nuts] AC calibration
>>
>> Adrian,
>> Do you have a link or title for the NIST paper that you mentioned?
>