Discussion of precise voltage measurement

View all threads

732A and Prologix received

D
Don@True-Cal
Tue, Aug 26, 2014 1:54 AM

Why?

Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.

BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.

This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.

Don Johnson

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received

On 8/25/2014 11:02 AM, Don@True-Cal wrote:

Silver or Gold plating on the terminal or wire will introduce the
undesirable  dissimilar metal properties, both at the plating junction
and at the plating metal to DUT terminal.

Why?

Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.

A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.

Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.

--
Mike


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Why? Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out. BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to. This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side. Don Johnson -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S Sent: Monday, August 25, 2014 11:03 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] 732A and Prologix received On 8/25/2014 11:02 AM, Don@True-Cal wrote: > Silver or Gold plating on the terminal or wire will introduce the > undesirable dissimilar metal properties, both at the plating junction > and at the plating metal to DUT terminal. Why? Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples. A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g. shortly after plugging in a banana plug warmed by body heat), they simply cancel. Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems. -- Mike _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
MS
Mike S
Tue, Aug 26, 2014 4:07 AM

On 8/25/2014 9:54 PM, Don@True-Cal wrote:

Why?

Let me count the ways.You can never count on any Seebeck voltage to
be immediately offset, there are far too many variables.

...

A set of
5440A-7002 (banana plug) cables comes with this calibrator
(5440A-7003 spade lugs for 5720A)

For those leads, Fluke says they have the advantage of "minimizing the
emf caused by temperature difference between the ends of the leads." I
don't have much argument with that - if there's a temperature difference
between the ends, you won't have an equal thermocouple offset, so
minimizing such effects makes sense.

But, since those leads use pure copper wiring ("Belden 8262 or 8719 or
equivalent"), how does having only the contact itself made of tellurium
copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas
tight crimped connections? Does the tinning on that copper cable have an
effect? What's the Seebeck coefficient for a copper to tellurium copper
connection?

You said that "Tellurium Copper is usually not used for a device's
terminal posts but used as the lead wire..." But, Fluke does exactly the
opposite - those Fluke leads use TC connectors and are characterized
"while engaged in a five-way binding post of Tellurium Copper Alloy 145,
half hard." Since they use pure copper cabling, I'd think they'd work
even better with a pure copper binding post, since the offsetting
thermocouples would be in close proximity, and not at different ends of
the cable.

You earlier said "Any type of Silver or Gold plating on the terminal or
wire will introduce the undesirable  dissimilar metal properties, both
at the plating junction and at the plating metal to DUT terminal." Now
you emphasize that by saying "If someone used a set of gold-plated
interconnects for this procedure, they would be laughed out of the lab
and the calibrator would be useless until recalibrated properly..."

But, the specs for the Fluke 5440A-7003 which you say are appropriate
for the 5720A state: "Connector materials: Copper space lug with gold
plating over silver plating." (Both platings!!) Is someone to be laughed
out of the lab for using the cables they're supposed to use? The
5440A-7002 specs aren't clear, but if they're not plated, are you
required to clean them of copper oxide for every use? It seems that
would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while
a copper-copper oxide one is ~1000 uV/C.

You seem to be stating one thing, but then giving evidence which
contradicts your claim. It seems that Fluke uses TC where there's
thermal mass, and pure copper where there's little, and uses plating in
addition.

Please clarify, this isn't making sense to me.

On 8/25/2014 9:54 PM, Don@True-Cal wrote: > Why? > > Let me count the ways.You can never count on any Seebeck voltage to > be immediately offset, there are far too many variables. ... > A set of > 5440A-7002 (banana plug) cables comes with this calibrator > (5440A-7003 spade lugs for 5720A) For those leads, Fluke says they have the advantage of "minimizing the emf caused by temperature difference between the ends of the leads." I don't have much argument with that - if there's a temperature difference between the ends, you won't have an equal thermocouple offset, so minimizing such effects makes sense. But, since those leads use pure copper wiring ("Belden 8262 or 8719 or equivalent"), how does having only the contact itself made of tellurium copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas tight crimped connections? Does the tinning on that copper cable have an effect? What's the Seebeck coefficient for a copper to tellurium copper connection? You said that "Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire..." But, Fluke does exactly the opposite - those Fluke leads use TC connectors and are characterized "while engaged in a five-way binding post of Tellurium Copper Alloy 145, half hard." Since they use pure copper cabling, I'd think they'd work even better with a pure copper binding post, since the offsetting thermocouples would be in close proximity, and not at different ends of the cable. You earlier said "Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal." Now you emphasize that by saying "If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly..." But, the specs for the Fluke 5440A-7003 which you say are appropriate for the 5720A state: "Connector materials: Copper space lug with gold plating over silver plating." (Both platings!!) Is someone to be laughed out of the lab for using the cables they're supposed to use? The 5440A-7002 specs aren't clear, but if they're not plated, are you required to clean them of copper oxide for every use? It seems that would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while a copper-copper oxide one is ~1000 uV/C. You seem to be stating one thing, but then giving evidence which contradicts your claim. It seems that Fluke uses TC where there's thermal mass, and pure copper where there's little, and uses plating in addition. Please clarify, this isn't making sense to me.
RE
Randy Evans
Tue, Aug 26, 2014 4:23 AM

Is there any way to tell when the function key routine is complete?  In the
case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete.  In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long while
before it's complete, but i have to guess when it's done.

Thanks,

Randy

On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:

Bill,

I am trying to figure out the MATH function without much success.  I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1.  I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display.  I tried it a couple of times and the
same result so I am doing something wrong.  Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.

Randy

On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:

Randy:

 The MATH function is accessible from the keypad.  I don't have an IEEE

interface right now that works.  You can also program the numeric keypad
keys to have preprogrammed functions.  DEFKEY

 I have made my own "low thermal" measurement leads from Pomona #4892

banana plugs and Belden #9272 wire.  Why 9272, because it was handy at the
time.  It is tin plated copper, shielded twisted pair 20 ga.  I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it.  I simply cannot find what I want so I will
build my own cable.  I have done something like this before and it worked
fine.  When I get a "round toit".

 I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have

used
in the past to verify my homemade "low thermal" leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.

 Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all

adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far
as
the instability of the readings it is hard to determine which is causing
the
problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
following code.  "MATH 14;NRDGS 40;TRIG 4;TRIG ;"  So what this does is
set
the MATH to "Statistics" (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to "40", the trigger to
"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button.  You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it.  This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering "MATH" and then a 2 for
low, a 4 for mean, and 13 for high.  Of course you could do all of this
through the IEEE also.  The 3458A has a very rich set of measurement
commands.  I am still learning all of them.  It depends upon what I am
trying to accomplish.

 Since the 1.018 and 1.0 volt outputs are passive and derived from

resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring.  I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved.  When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts.  This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings.  So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above.  I
would
use this to determine where your problem might exist.  Just having the
meter
input shorted will point you in the right direction.  Meter, cables or
732A.

 Sorry for the long dissertation.  Friends get mad at me for being so

detailed sometimes.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received

Bill,

I have convinced myself that the problem I an seeing is due to thermals.
If I move the cables (with gold-plated banana plugs) using a small

towel

rather than letting my hand touch the plugs, it is much more stable.

If I

then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward.  I am trying to check the stability

of

the reading but I haven't figured out the MATH function yet.  I assume

this

is a programmed function using GPIB only?

The stability I am seeing by manually recording readings (using NLPC of

100

and 1000) is much greater than what you are measuring on your system.

Not

sure how to ascertain if it's the 3458A or the 732.  The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output.  Rather

large

differences (this is after an ACAL).  I need to find some better cables

to

make sure the errors are not due to thermals again.

Randy

On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net wrote:

Randy:

 I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit

perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
others
that are in this size and package.  Order from one of the usual

electronics

distributors like Allied, Mouser, Digikey.  This is a very common

battery

as
it is used in a lot of "EXIT" signs so they are lighted when the power

goes

out.  I don't see how the 12V 7AH will fit as they are too large.  I

guess

you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt

4
AH

but the terminals are in the wrong place so you will have to "nibble"

out

the aluminum plate that holds them in the 732A battery pack.  You have

to

be
careful if you use the 12v 5AH as you will have 4 extra battery

connection

leads to deal with and connect correctly.  I would stick with the 6V

4AH.

New batteries will last around 12 to 14 hours before the "CAL" light

goes

out when AC power is not applied.  So shipping to Cal Lab can be a

problem

if it is a distance away, or you have to use a shipper like UPS or

FEDEX

and
you ship the night before and then use their "Morning delivery" and

the
Cal

Lab is expecting your 732A.  Same on the way back to you.  Of course

you

could always strap another battery on the 732A and hook it up to the

"ext

power" plug to last longer.  I have seen it done.  The issue is to get

the

Cal Lab to charge the extra battery before they ship the 732A back to

you.

 When you remove and work on the battery pack always have the AC

power

plugged in.  The "CAL" led will stay on because the 18.6 v regulated

supply

is working.

 The "CAL" light is to indicate that power has not been lost to the

Reference Amp or other associated circuits.  When the raw supply

(battery)

voltage drops below about 21 volts the "CAL" light will go out.  Below

that

voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate.  The requirement is that the

Reference

Amp be kept "alive" at all times to maintain the output voltage that

was

measured at the time of the most recent Calibration or Certification.

When

the semiconductor junctions are unbiased and cool off when power is

lost,

and then power is restored the result will be a different 10 volts

than

before the power failure.  My experience is that after all of the

years

that
these units have been powered up, this won't happen and when power is

lost

and then restored, even months later, the 732A will come back to

almost

exactly the same 10 volts as when they lost power, usually with in 0.2

PPM

after 24 hours of "warm up".

 What type of hookup leads are you using when measuring the 1 volt

output?  What is the PLC set to?  I always use 100 PLC to measure

this.
If

you don't have "low thermal" connection leads you can experience uV

changes

for a minute or more after plugging in the leads due to the "thermals"
generated because of the difference in temperature between the banana

jacks

on the 732A and the banana plugs of the leads.  I have found that even

just

plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my

theory

at any rate).  You have to allow at least a minute or more before

being

able
to make a measurement after plugging in the leads.  I just measured

the

variation of the 1 volt output of my 732A and using my 3458A and I

got a

total difference of  0.159 uV over 40 measurements using 100 PLC on

the
1

volt range of the 3458A.  Using the MATH function and all of the data

you

can collect.  That was after waiting for several minutes after

plugging
in

the leads.

 I hope all of this helps.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received

Todd,

Thanks for the info.  I have several Panasonic 12V 7 AH batteries

that
I

keep topped off and they have very low current draw (~2 to 3 mA at

13.5

VDC) when charged and at their float voltage, so I am pretty sure

they

are

in good condition.  I will look at getting those in the units after

I

ascertain the condition of the 732.

So now I have a what appears to be a functioning 3458A and a 732A

but

they

slightly disagree.  I am like the man with two watches that disagree

on

the

time  - which is correct?  For the moment, i am only concerned with
stability.  The need for absolute accuracy will come later.

Randy

On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com

wrote:

Randy,

You have two possible choices. It can be configured with 4 x 6v

4Ah

batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner

has

modified the battery pack already. A couple of mine needed a

nibbler

tool

to remove enough of the aluminum cover that fits over the tops of

the

batteries. The original cover will short out to the battery tabs

regardless

of the battery configuration if this is not done.

You can find larger capacity batteries that will give you slightly

more

battery life. I lost a couple sets of mail-order batteries after a

few

extended outages. I would recommend going with locally bought

batteries

instead of the cheaper mail order. My local Batteries Plus will

typically

have some warranty if I remember correctly. Moving forward I will

only

use

2 12v batteries and pre-charge them on a battery charger to

equalize

them

before putting them in the 732A. I think the cheap batteries did

not

discharge equally, and would not recover when power was applied.

Inspect the back plane for damaged traces and look at the

capacitors. I

had

a few that looked questionable. So far, I have replaced all the

big

caps
on

the pre-regulator and regulator boards. My feeling is that once

these

go

online, they should run as long as possible between repairs.

The battery charger circuit may need adjusting. I tweaked mine and

it

seemed to work fine.

Todd

On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <

wrote:

I received my Fluke 732A today.  Just powered it up but it needs

new

batteries.  Any suggestions for sources (I haven't opened up the

unit

yet -

I want to make sure it works before doing that).  Also received

the

ProLogix USB-GPIB adapter.  I plan on using Mark Sims' CAL ran

data

dumper

program to get the CAL data from my 3458A.  Should be a busy

weekend.

Randy


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Is there any way to tell when the function key routine is complete? In the case of taking multiple readings using the DEFKEY and MATH function, I don't see any indication when the routine is complete. In one particular case, I am taking a 100 readings with NLPC set for 1000 so its a long while before it's complete, but i have to guess when it's done. Thanks, Randy On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com> wrote: > Bill, > > I am trying to figure out the MATH function without much success. I input > the sequence you said (I looked up the instructions to understand what you > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and > it shows up on the display when I input BLUE F1. I hit ENTER and it takes > the 40 measurements and the MATH symbol shows on the display during the > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I > get a MATH ERR symbol on the display. I tried it a couple of times and the > same result so I am doing something wrong. Is there a better source for > explaining how to do front panel masurements than the User Guide, which > seems oriented at programming automatic rather than manual measurements. > > Randy > > > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote: > >> Randy: >> >> The MATH function is accessible from the keypad. I don't have an IEEE >> interface right now that works. You can also program the numeric keypad >> keys to have preprogrammed functions. DEFKEY >> >> I have made my own "low thermal" measurement leads from Pomona #4892 >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the >> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans >> to do custom cables with 16 ga. bare copper wire that I will twist and >> then >> put a braided shield over it. I simply cannot find what I want so I will >> build my own cable. I have done something like this before and it worked >> fine. When I get a "round toit". >> >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have >> used >> in the past to verify my homemade "low thermal" leads as described above. >> Frankly I cannot see any difference between using the 1756 cables and my >> homemade cables once I give them a few minutes for the thermals to go >> away. >> As far as I can tell and measure the differences, if any, are below 0.1 >> ppm >> at 10 volts. >> >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far >> as >> the instability of the readings it is hard to determine which is causing >> the >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is >> set >> the MATH to "Statistics" (store high reading/low reading/ and mean of the >> readings) in the registers, the number of readings to "40", the trigger to >> "hold" (which keeps the meter from triggering until I press "ENTER" and >> then >> trigger the sequence of 40 readings when I push the "ENTER" button. You >> can >> do all of this manually from the keypads but since I use this sequence a >> lot >> I have preprogrammed it. This is after I set digits to "8" and PLC to >> 100. >> Once those 40 readings are finished then you can access the various MATH >> statistic registers, using the menu, by entering "MATH" and then a 2 for >> low, a 4 for mean, and 13 for high. Of course you could do all of this >> through the IEEE also. The 3458A has a very rich set of measurement >> commands. I am still learning all of them. It depends upon what I am >> trying to accomplish. >> >> Since the 1.018 and 1.0 volt outputs are passive and derived from >> resistive dividers from the 10 volt, I don't see how they could contribute >> to the varying readings you are measuring. I think I would put a short on >> the input of the 3458A and manually set the range to 1 volt and then >> observe >> the variations that way without the 732A involved. When I do this I see a >> variation from low reading to high reading of 0.125 uVolts and then >> another >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low >> side >> of the measurment terminals, GUARD connected doesn't seem to affect the >> readings. So that is the base noise of the 3458A without the 732A, >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I >> got >> a variation of 0.159 uVolts using the same 40 reading method above. I >> would >> use this to determine where your problem might exist. Just having the >> meter >> input shorted will point you in the right direction. Meter, cables or >> 732A. >> >> Sorry for the long dissertation. Friends get mad at me for being so >> detailed sometimes. >> >> Bill >> >> ----- Original Message ----- >> From: "Randy Evans" <randyevans2688@gmail.com> >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> Sent: Sunday, August 24, 2014 7:22 AM >> Subject: Re: [volt-nuts] 732A and Prologix received >> >> >> > Bill, >> > >> > I have convinced myself that the problem I an seeing is due to thermals. >> > If I move the cables (with gold-plated banana plugs) using a small >> towel >> > rather than letting my hand touch the plugs, it is much more stable. >> If I >> > then hold the banana plug with my hand after the reading has stabilized, >> > the reading drifts rapidly upward. I am trying to check the stability >> of >> > the reading but I haven't figured out the MATH function yet. I assume >> this >> > is a programmed function using GPIB only? >> > >> > The stability I am seeing by manually recording readings (using NLPC of >> 100 >> > and 1000) is much greater than what you are measuring on your system. >> Not >> > sure how to ascertain if it's the 3458A or the 732. The value of the >> > readings are very different between the two - the 3458 reads about 50 uV >> > high on the 10 V output and about 12 uV low on the 1V output. Rather >> large >> > differences (this is after an ACAL). I need to find some better cables >> to >> > make sure the errors are not due to thermals again. >> > >> > Randy >> > >> > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote: >> > >> > > Randy: >> > > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and >> > > others >> > > that are in this size and package. Order from one of the usual >> electronics >> > > distributors like Allied, Mouser, Digikey. This is a very common >> battery >> > > as >> > > it is used in a lot of "EXIT" signs so they are lighted when the power >> goes >> > > out. I don't see how the 12V 7AH will fit as they are too large. I >> guess >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt >> 4 >> AH >> > > but the terminals are in the wrong place so you will have to "nibble" >> out >> > > the aluminum plate that holds them in the 732A battery pack. You have >> to >> > > be >> > > careful if you use the 12v 5AH as you will have 4 extra battery >> connection >> > > leads to deal with and connect correctly. I would stick with the 6V >> 4AH. >> > > New batteries will last around 12 to 14 hours before the "CAL" light >> goes >> > > out when AC power is not applied. So shipping to Cal Lab can be a >> problem >> > > if it is a distance away, or you have to use a shipper like UPS or >> FEDEX >> > > and >> > > you ship the night before and then use their "Morning delivery" and >> the >> Cal >> > > Lab is expecting your 732A. Same on the way back to you. Of course >> you >> > > could always strap another battery on the 732A and hook it up to the >> "ext >> > > power" plug to last longer. I have seen it done. The issue is to get >> the >> > > Cal Lab to charge the extra battery before they ship the 732A back to >> you. >> > > >> > > When you remove and work on the battery pack always have the AC >> power >> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated >> supply >> > > is working. >> > > >> > > The "CAL" light is to indicate that power has not been lost to the >> > > Reference Amp or other associated circuits. When the raw supply >> (battery) >> > > voltage drops below about 21 volts the "CAL" light will go out. Below >> that >> > > voltage the heater circuits will not work correctly and the 18.6 volt >> > > regulated supply will not regulate. The requirement is that the >> Reference >> > > Amp be kept "alive" at all times to maintain the output voltage that >> was >> > > measured at the time of the most recent Calibration or Certification. >> When >> > > the semiconductor junctions are unbiased and cool off when power is >> lost, >> > > and then power is restored the result will be a different 10 volts >> than >> > > before the power failure. My experience is that after all of the >> years >> > > that >> > > these units have been powered up, this won't happen and when power is >> lost >> > > and then restored, even months later, the 732A will come back to >> almost >> > > exactly the same 10 volts as when they lost power, usually with in 0.2 >> PPM >> > > after 24 hours of "warm up". >> > > >> > > What type of hookup leads are you using when measuring the 1 volt >> > > output? What is the PLC set to? I always use 100 PLC to measure >> this. >> If >> > > you don't have "low thermal" connection leads you can experience uV >> changes >> > > for a minute or more after plugging in the leads due to the "thermals" >> > > generated because of the difference in temperature between the banana >> jacks >> > > on the 732A and the banana plugs of the leads. I have found that even >> just >> > > plugging in the lead will generate a thermal difference because of >> > > difference of temps and some heating due to the physical act of just >> > > inserting the plug because of friction between the jack and plug (my >> theory >> > > at any rate). You have to allow at least a minute or more before >> being >> > > able >> > > to make a measurement after plugging in the leads. I just measured >> the >> > > variation of the 1 volt output of my 732A and using my 3458A and I >> got a >> > > total difference of 0.159 uV over 40 measurements using 100 PLC on >> the >> 1 >> > > volt range of the 3458A. Using the MATH function and all of the data >> you >> > > can collect. That was after waiting for several minutes after >> plugging >> in >> > > the leads. >> > > >> > > I hope all of this helps. >> > > >> > > Bill >> > > >> > > >> > > >> > > ----- Original Message ----- >> > > From: "Randy Evans" <randyevans2688@gmail.com> >> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> > > Sent: Friday, August 22, 2014 9:03 PM >> > > Subject: Re: [volt-nuts] 732A and Prologix received >> > > >> > > >> > > > Todd, >> > > > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries >> that >> I >> > > > keep topped off and they have very low current draw (~2 to 3 mA at >> 13.5 >> > > > VDC) when charged and at their float voltage, so I am pretty sure >> they >> > > are >> > > > in good condition. I will look at getting those in the units after >> I >> > > > ascertain the condition of the 732. >> > > > >> > > > So now I have a what appears to be a functioning 3458A and a 732A >> but >> > > they >> > > > slightly disagree. I am like the man with two watches that disagree >> on >> > > the >> > > > time - which is correct? For the moment, i am only concerned with >> > > > stability. The need for absolute accuracy will come later. >> > > > >> > > > Randy >> > > > >> > > > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com >> > >> > > wrote: >> > > > >> > > > > Randy, >> > > > > >> > > > > You have two possible choices. It can be configured with 4 x 6v >> 4Ah >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner >> has >> > > > > modified the battery pack already. A couple of mine needed a >> nibbler >> > > tool >> > > > > to remove enough of the aluminum cover that fits over the tops of >> the >> > > > > batteries. The original cover will short out to the battery tabs >> > > regardless >> > > > > of the battery configuration if this is not done. >> > > > > >> > > > > You can find larger capacity batteries that will give you slightly >> more >> > > > > battery life. I lost a couple sets of mail-order batteries after a >> few >> > > > > extended outages. I would recommend going with locally bought >> batteries >> > > > > instead of the cheaper mail order. My local Batteries Plus will >> > > typically >> > > > > have some warranty if I remember correctly. Moving forward I will >> only >> > > use >> > > > > 2 12v batteries and pre-charge them on a battery charger to >> equalize >> > > them >> > > > > before putting them in the 732A. I think the cheap batteries did >> not >> > > > > discharge equally, and would not recover when power was applied. >> > > > > >> > > > > Inspect the back plane for damaged traces and look at the >> capacitors. I >> > > had >> > > > > a few that looked questionable. So far, I have replaced all the >> big >> > > caps >> > > on >> > > > > the pre-regulator and regulator boards. My feeling is that once >> these >> > > go >> > > > > online, they should run as long as possible between repairs. >> > > > > >> > > > > The battery charger circuit may need adjusting. I tweaked mine and >> it >> > > > > seemed to work fine. >> > > > > >> > > > > Todd >> > > > > >> > > > > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans < >> > > randyevans2688@gmail.com> >> > > > > wrote: >> > > > > >> > > > > > I received my Fluke 732A today. Just powered it up but it needs >> new >> > > > > > batteries. Any suggestions for sources (I haven't opened up the >> unit >> > > > > yet - >> > > > > > I want to make sure it works before doing that). Also received >> the >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran >> data >> > > > > dumper >> > > > > > program to get the CAL data from my 3458A. Should be a busy >> weekend. >> > > > > > >> > > > > > Randy >> > > > > > _______________________________________________ >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com >> > > > > > To unsubscribe, go to >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> > > > > > and follow the instructions there. >> > > > > > >> > > > > _______________________________________________ >> > > > > volt-nuts mailing list -- volt-nuts@febo.com >> > > > > To unsubscribe, go to >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> > > > > and follow the instructions there. >> > > > > >> > > > _______________________________________________ >> > > > volt-nuts mailing list -- volt-nuts@febo.com >> > > > To unsubscribe, go to >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> > > > and follow the instructions there. >> > > >> > > _______________________________________________ >> > > volt-nuts mailing list -- volt-nuts@febo.com >> > > To unsubscribe, go to >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> > > and follow the instructions there. >> > > >> > _______________________________________________ >> > volt-nuts mailing list -- volt-nuts@febo.com >> > To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> > and follow the instructions there. >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > >
CB
Charles Black
Tue, Aug 26, 2014 4:28 AM

My post on this thread earlier today seems to be lost (maybe the system
change) so I am re-posting some of it (corrected) since it's apropos to
the discussion.

Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire. This is very convenient since it is so
easy to duplicate in the field and makes simple inexpensive test leads
best for high precision measurements. In order for the 3458A to make
full accuracy measurements (8 digit) NPLC must be set to 1000 (according
to the User's Guide). If you use a lower NPLC value there is a table in
the User's Guide that can be used to determine how accurate your
measurements are going to be.

Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.

Calibration shunt            -0.00021mv ± 10nV  Equilibration time 5
minutes. 14 gauge per Calibration Manual.
"U" heavy wire              -0.00021mv ± 10nV  Equil. time 5 minutes.
Used the through holes in the Input banana posts only.
Copper wire                    -0.00019mv ± 10nV  NAPA PVC covered
automobile wire at same contact points as CAL shunt
Standard Ground Plate  -0.00040mv ± 10nV  Equil. time 5 minutes. Gold
plated ground plate from my Datron 4910
Copper wire                    -0.00019mv ± 10nv  Equil. time 2
seconds. Used the banana through holes.
STP 2 meter test lead    -0.00021mv ± 10nV  Equil. time 2 seconds.
M27500 24 gauge STP Tefzel insulation.
Banana plugs                  -0.00021mv ± 30nV  Equil. time 20
minutes. My best "Perfect" gold plated plugs with copper wire.

Note that all my test shorts equilibrate to virtualy zero volts. There
is almost no Seebeck voltage for any of my test shorts as long as the
system is allowed to equilibrate. Equilibation times varied between less
than 2 seconds up 20 minutes with longer times due to excessive thermal
mass. Excessive thermal mass also caused some voltage instability.

Charlie

On 8/25/2014 6:54 PM, Don@True-Cal wrote:

Why?

Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out.

BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to.

This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side.

Don Johnson

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received

On 8/25/2014 11:02 AM, Don@True-Cal wrote:

Silver or Gold plating on the terminal or wire will introduce the
undesirable  dissimilar metal properties, both at the plating junction
and at the plating metal to DUT terminal.

Why?

Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples.

A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply cancel.

Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems.

--
Mike


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

My post on this thread earlier today seems to be lost (maybe the system change) so I am re-posting some of it (corrected) since it's apropos to the discussion. Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information to deduce that lowest input short voltage is going to be a copper wire since the meter is set to zero volts during calibration using a heavy gauge (14 to 16) copper wire. This is very convenient since it is so easy to duplicate in the field and makes simple inexpensive test leads best for high precision measurements. In order for the 3458A to make full accuracy measurements (8 digit) NPLC must be set to 1000 (according to the User's Guide). If you use a lower NPLC value there is a table in the User's Guide that can be used to determine how accurate your measurements are going to be. Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been about a year since I clid my last CAL 0 so it was going to be interesting at least for me. The first shunt was my test "U' shaped shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair. Calibration shunt -0.00021mv ± 10nV Equilibration time 5 minutes. 14 gauge per Calibration Manual. "U" heavy wire -0.00021mv ± 10nV Equil. time 5 minutes. Used the through holes in the Input banana posts only. Copper wire -0.00019mv ± 10nV NAPA PVC covered automobile wire at same contact points as CAL shunt Standard Ground Plate -0.00040mv ± 10nV Equil. time 5 minutes. Gold plated ground plate from my Datron 4910 Copper wire -0.00019mv ± 10nv Equil. time 2 seconds. Used the banana through holes. STP 2 meter test lead -0.00021mv ± 10nV Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation. Banana plugs -0.00021mv ± 30nV Equil. time 20 minutes. My best "Perfect" gold plated plugs with copper wire. Note that all my test shorts equilibrate to virtualy zero volts. There is almost no Seebeck voltage for any of my test shorts as long as the system is allowed to equilibrate. Equilibation times varied between less than 2 seconds up 20 minutes with longer times due to excessive thermal mass. Excessive thermal mass also caused some voltage instability. Charlie On 8/25/2014 6:54 PM, Don@True-Cal wrote: > Why? > > Let me count the ways. You can never count on any Seebeck voltage to be immediately offset, there are far too many variables. Best example I can think of...why is there an Ohms Offset Compensation feature on any good high resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 4-wire mode using inexpensive nickel-plated leads and even allow plenty of time for everything to thermally stabilize. Using Ohms Offset Compensation, enable and disable it and observe the difference. If the Seebeck voltages were all immediately offset, as you say, there would be no difference. But there most certainly is. Or simply, why is there a need for ohms offset compensation feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a horrible choice but if it all canceled, what difference would it make how bad is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips back to Fluke for full calibration, there is an interim external calibration procedure using the 732B, 742A-1 & 742A-10k. If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly. A set of 5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade lugs for 5720A) and recommended for the calibration procedure but other Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts will never allow true thermal symmetry around the DMM or DUT terminals. To convince yourself, place an oscillating fan several feet back from the DMM and DUT terminals and using the 1 or 10 Ohm setup from above, again with the nickel-plated leads, watch the variations. Sure the fan and the nickel-plated exaggerates the issue but it quickly dispels the notion that all the Seebeck voltages are canceled out. > > BTW, the plating layer temperature on a plated terminal will be somewhere between the temperature of the base metal and mating terminal it's connected to. > > This is not just theory, my 40+ years in the cal lab is driving my arguments but it never hurts to have physical-science on your side. > > Don Johnson > > -----Original Message----- > From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mike S > Sent: Monday, August 25, 2014 11:03 AM > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] 732A and Prologix received > > On 8/25/2014 11:02 AM, Don@True-Cal wrote: >> Silver or Gold plating on the terminal or wire will introduce the >> undesirable dissimilar metal properties, both at the plating junction >> and at the plating metal to DUT terminal. > Why? > > Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples. > > A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g. > shortly after plugging in a banana plug warmed by body heat), they simply cancel. > > Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems. > > -- > Mike > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > >
BG
Bill Gold
Tue, Aug 26, 2014 5:01 AM

Not that I know of.  Just wait and when you don't see a "SMPL" on the
display it is done.  But then with 1000 PLC that is around 16.66 seconds per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes.  Not from the front panel
at any rate.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received

Is there any way to tell when the function key routine is complete?  In

the

case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete.  In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long

while

before it's complete, but i have to guess when it's done.

Thanks,

Randy

On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:

Bill,

I am trying to figure out the MATH function without much success.  I

input

the sequence you said (I looked up the instructions to understand what

you

did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;

and

it shows up on the display when I input BLUE F1.  I hit ENTER and it

takes

the 40 measurements and the MATH symbol shows on the display during the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display.  I tried it a couple of times and

the

same result so I am doing something wrong.  Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.

Randy

On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:

Randy:

 The MATH function is accessible from the keypad.  I don't have an

IEEE

interface right now that works.  You can also program the numeric

keypad

keys to have preprogrammed functions.  DEFKEY

 I have made my own "low thermal" measurement leads from Pomona

#4892

banana plugs and Belden #9272 wire.  Why 9272, because it was handy at

the

time.  It is tin plated copper, shielded twisted pair 20 ga.  I have

plans

to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it.  I simply cannot find what I want so I

will

build my own cable.  I have done something like this before and it

worked

fine.  When I get a "round toit".

 I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have

used
in the past to verify my homemade "low thermal" leads as described

above.

Frankly I cannot see any difference between using the 1756 cables and

my

homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below 0.1
ppm
at 10 volts.

 Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are

all

adjustable you may be seeing a misadjusted 1 volt from the 732A.  As

far

as
the instability of the readings it is hard to determine which is

causing

the
problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
following code.  "MATH 14;NRDGS 40;TRIG 4;TRIG ;"  So what this does

is

set
the MATH to "Statistics" (store high reading/low reading/ and mean of

the

readings) in the registers, the number of readings to "40", the trigger

to

"hold" (which keeps the meter from triggering until I press "ENTER" and
then
trigger the sequence of 40 readings when I push the "ENTER" button.

You

can
do all of this manually from the keypads but since I use this sequence

a

lot
I have preprogrammed it.  This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various

MATH

statistic registers, using the menu, by entering "MATH" and then a 2

for

low, a 4 for mean, and 13 for high.  Of course you could do all of this
through the IEEE also.  The 3458A has a very rich set of measurement
commands.  I am still learning all of them.  It depends upon what I am
trying to accomplish.

 Since the 1.018 and 1.0 volt outputs are passive and derived from

resistive dividers from the 10 volt, I don't see how they could

contribute

to the varying readings you are measuring.  I think I would put a short

on

the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved.  When I do this I

see a

variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts.  This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings.  So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I
got
a variation of 0.159 uVolts using the same 40 reading method above.  I
would
use this to determine where your problem might exist.  Just having the
meter
input shorted will point you in the right direction.  Meter, cables or
732A.

 Sorry for the long dissertation.  Friends get mad at me for being

so

detailed sometimes.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received

Bill,

I have convinced myself that the problem I an seeing is due to

thermals.

If I move the cables (with gold-plated banana plugs) using a small

towel

rather than letting my hand touch the plugs, it is much more stable.

If I

then hold the banana plug with my hand after the reading has

stabilized,

the reading drifts rapidly upward.  I am trying to check the

stability

of

the reading but I haven't figured out the MATH function yet.  I

assume

this

is a programmed function using GPIB only?

The stability I am seeing by manually recording readings (using NLPC

of

100

and 1000) is much greater than what you are measuring on your system.

Not

sure how to ascertain if it's the 3458A or the 732.  The value of the
readings are very different between the two - the 3458 reads about 50

uV

high on the 10 V output and about 12 uV low on the 1V output.  Rather

large

differences (this is after an ACAL).  I need to find some better

cables

to

make sure the errors are not due to thermals again.

Randy

On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net

wrote:

Randy:

 I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will

fit

perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C

and

others
that are in this size and package.  Order from one of the usual

electronics

distributors like Allied, Mouser, Digikey.  This is a very common

battery

as
it is used in a lot of "EXIT" signs so they are lighted when the

power

goes

out.  I don't see how the 12V 7AH will fit as they are too large.

I

guess

you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6

volt

4
AH

but the terminals are in the wrong place so you will have to

"nibble"

out

the aluminum plate that holds them in the 732A battery pack.  You

have

to

be
careful if you use the 12v 5AH as you will have 4 extra battery

connection

leads to deal with and connect correctly.  I would stick with the

6V

4AH.

New batteries will last around 12 to 14 hours before the "CAL"

light

goes

out when AC power is not applied.  So shipping to Cal Lab can be a

problem

if it is a distance away, or you have to use a shipper like UPS or

FEDEX

and
you ship the night before and then use their "Morning delivery" and

the
Cal

Lab is expecting your 732A.  Same on the way back to you.  Of

course

you

could always strap another battery on the 732A and hook it up to

the

"ext

power" plug to last longer.  I have seen it done.  The issue is to

get

the

Cal Lab to charge the extra battery before they ship the 732A back

to

you.

 When you remove and work on the battery pack always have the AC

power

plugged in.  The "CAL" led will stay on because the 18.6 v

regulated

supply

is working.

 The "CAL" light is to indicate that power has not been lost to

the

Reference Amp or other associated circuits.  When the raw supply

(battery)

voltage drops below about 21 volts the "CAL" light will go out.

Below

that

voltage the heater circuits will not work correctly and the 18.6

volt

regulated supply will not regulate.  The requirement is that the

Reference

Amp be kept "alive" at all times to maintain the output voltage

that

was

measured at the time of the most recent Calibration or

Certification.

When

the semiconductor junctions are unbiased and cool off when power is

lost,

and then power is restored the result will be a different 10 volts

than

before the power failure.  My experience is that after all of the

years

that
these units have been powered up, this won't happen and when power

is

lost

and then restored, even months later, the 732A will come back to

almost

exactly the same 10 volts as when they lost power, usually with in

0.2

PPM

after 24 hours of "warm up".

 What type of hookup leads are you using when measuring the 1

volt

output?  What is the PLC set to?  I always use 100 PLC to measure

this.
If

you don't have "low thermal" connection leads you can experience uV

changes

for a minute or more after plugging in the leads due to the

"thermals"

generated because of the difference in temperature between the

banana

jacks

on the 732A and the banana plugs of the leads.  I have found that

even

just

plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of

just

inserting the plug because of friction between the jack and plug

(my

theory

at any rate).  You have to allow at least a minute or more before

being

able
to make a measurement after plugging in the leads.  I just measured

the

variation of the 1 volt output of my 732A and using my 3458A and I

got a

total difference of  0.159 uV over 40 measurements using 100 PLC on

the
1

volt range of the 3458A.  Using the MATH function and all of the

data

you

can collect.  That was after waiting for several minutes after

plugging
in

the leads.

 I hope all of this helps.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"

Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received

Todd,

Thanks for the info.  I have several Panasonic 12V 7 AH batteries

that
I

keep topped off and they have very low current draw (~2 to 3 mA

at

13.5

VDC) when charged and at their float voltage, so I am pretty sure

they

are

in good condition.  I will look at getting those in the units

after

I

ascertain the condition of the 732.

So now I have a what appears to be a functioning 3458A and a 732A

but

they

slightly disagree.  I am like the man with two watches that

disagree

on

the

time  - which is correct?  For the moment, i am only concerned

with

stability.  The need for absolute accuracy will come later.

Randy

On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef

wrote:

Randy,

You have two possible choices. It can be configured with 4 x 6v

4Ah

batteries or 2 x 12v 7Ah batteries. Hopefully the previous

owner

has

modified the battery pack already. A couple of mine needed a

nibbler

tool

to remove enough of the aluminum cover that fits over the tops

of

the

batteries. The original cover will short out to the battery

tabs

regardless

of the battery configuration if this is not done.

You can find larger capacity batteries that will give you

slightly

more

battery life. I lost a couple sets of mail-order batteries

after a

few

extended outages. I would recommend going with locally bought

batteries

instead of the cheaper mail order. My local Batteries Plus will

typically

have some warranty if I remember correctly. Moving forward I

will

only

use

2 12v batteries and pre-charge them on a battery charger to

equalize

them

before putting them in the 732A. I think the cheap batteries

did

not

discharge equally, and would not recover when power was

applied.

Inspect the back plane for damaged traces and look at the

capacitors. I

had

a few that looked questionable. So far, I have replaced all the

big

caps
on

the pre-regulator and regulator boards. My feeling is that once

these

go

online, they should run as long as possible between repairs.

The battery charger circuit may need adjusting. I tweaked mine

and

it

seemed to work fine.

Todd

On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <

wrote:

I received my Fluke 732A today.  Just powered it up but it

needs

new

batteries.  Any suggestions for sources (I haven't opened up

the

unit

yet -

I want to make sure it works before doing that).  Also

received

the

ProLogix USB-GPIB adapter.  I plan on using Mark Sims' CAL

ran

data

dumper

program to get the CAL data from my 3458A.  Should be a busy

weekend.

Randy


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.

Not that I know of. Just wait and when you don't see a "SMPL" on the display it is done. But then with 1000 PLC that is around 16.66 seconds per reading times 100 readings is somewhere around 28 minutes and there is probably some overhead time so around 30 minutes. Not from the front panel at any rate. Bill ----- Original Message ----- From: "Randy Evans" <randyevans2688@gmail.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Monday, August 25, 2014 9:23 PM Subject: Re: [volt-nuts] 732A and Prologix received > Is there any way to tell when the function key routine is complete? In the > case of taking multiple readings using the DEFKEY and MATH function, I > don't see any indication when the routine is complete. In one particular > case, I am taking a 100 readings with NLPC set for 1000 so its a long while > before it's complete, but i have to guess when it's done. > > Thanks, > > Randy > > > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com> > wrote: > > > Bill, > > > > I am trying to figure out the MATH function without much success. I input > > the sequence you said (I looked up the instructions to understand what you > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and > > it shows up on the display when I input BLUE F1. I hit ENTER and it takes > > the 40 measurements and the MATH symbol shows on the display during the > > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I > > get a MATH ERR symbol on the display. I tried it a couple of times and the > > same result so I am doing something wrong. Is there a better source for > > explaining how to do front panel masurements than the User Guide, which > > seems oriented at programming automatic rather than manual measurements. > > > > Randy > > > > > > > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote: > > > >> Randy: > >> > >> The MATH function is accessible from the keypad. I don't have an IEEE > >> interface right now that works. You can also program the numeric keypad > >> keys to have preprogrammed functions. DEFKEY > >> > >> I have made my own "low thermal" measurement leads from Pomona #4892 > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans > >> to do custom cables with 16 ga. bare copper wire that I will twist and > >> then > >> put a braided shield over it. I simply cannot find what I want so I will > >> build my own cable. I have done something like this before and it worked > >> fine. When I get a "round toit". > >> > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have > >> used > >> in the past to verify my homemade "low thermal" leads as described above. > >> Frankly I cannot see any difference between using the 1756 cables and my > >> homemade cables once I give them a few minutes for the thermals to go > >> away. > >> As far as I can tell and measure the differences, if any, are below 0.1 > >> ppm > >> at 10 volts. > >> > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far > >> as > >> the instability of the readings it is hard to determine which is causing > >> the > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is > >> set > >> the MATH to "Statistics" (store high reading/low reading/ and mean of the > >> readings) in the registers, the number of readings to "40", the trigger to > >> "hold" (which keeps the meter from triggering until I press "ENTER" and > >> then > >> trigger the sequence of 40 readings when I push the "ENTER" button. You > >> can > >> do all of this manually from the keypads but since I use this sequence a > >> lot > >> I have preprogrammed it. This is after I set digits to "8" and PLC to > >> 100. > >> Once those 40 readings are finished then you can access the various MATH > >> statistic registers, using the menu, by entering "MATH" and then a 2 for > >> low, a 4 for mean, and 13 for high. Of course you could do all of this > >> through the IEEE also. The 3458A has a very rich set of measurement > >> commands. I am still learning all of them. It depends upon what I am > >> trying to accomplish. > >> > >> Since the 1.018 and 1.0 volt outputs are passive and derived from > >> resistive dividers from the 10 volt, I don't see how they could contribute > >> to the varying readings you are measuring. I think I would put a short on > >> the input of the 3458A and manually set the range to 1 volt and then > >> observe > >> the variations that way without the 732A involved. When I do this I see a > >> variation from low reading to high reading of 0.125 uVolts and then > >> another > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low > >> side > >> of the measurment terminals, GUARD connected doesn't seem to affect the > >> readings. So that is the base noise of the 3458A without the 732A, > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I > >> got > >> a variation of 0.159 uVolts using the same 40 reading method above. I > >> would > >> use this to determine where your problem might exist. Just having the > >> meter > >> input shorted will point you in the right direction. Meter, cables or > >> 732A. > >> > >> Sorry for the long dissertation. Friends get mad at me for being so > >> detailed sometimes. > >> > >> Bill > >> > >> ----- Original Message ----- > >> From: "Randy Evans" <randyevans2688@gmail.com> > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > >> Sent: Sunday, August 24, 2014 7:22 AM > >> Subject: Re: [volt-nuts] 732A and Prologix received > >> > >> > >> > Bill, > >> > > >> > I have convinced myself that the problem I an seeing is due to thermals. > >> > If I move the cables (with gold-plated banana plugs) using a small > >> towel > >> > rather than letting my hand touch the plugs, it is much more stable. > >> If I > >> > then hold the banana plug with my hand after the reading has stabilized, > >> > the reading drifts rapidly upward. I am trying to check the stability > >> of > >> > the reading but I haven't figured out the MATH function yet. I assume > >> this > >> > is a programmed function using GPIB only? > >> > > >> > The stability I am seeing by manually recording readings (using NLPC of > >> 100 > >> > and 1000) is much greater than what you are measuring on your system. > >> Not > >> > sure how to ascertain if it's the 3458A or the 732. The value of the > >> > readings are very different between the two - the 3458 reads about 50 uV > >> > high on the 10 V output and about 12 uV low on the 1V output. Rather > >> large > >> > differences (this is after an ACAL). I need to find some better cables > >> to > >> > make sure the errors are not due to thermals again. > >> > > >> > Randy > >> > > >> > > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> wrote: > >> > > >> > > Randy: > >> > > > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and > >> > > others > >> > > that are in this size and package. Order from one of the usual > >> electronics > >> > > distributors like Allied, Mouser, Digikey. This is a very common > >> battery > >> > > as > >> > > it is used in a lot of "EXIT" signs so they are lighted when the power > >> goes > >> > > out. I don't see how the 12V 7AH will fit as they are too large. I > >> guess > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt > >> 4 > >> AH > >> > > but the terminals are in the wrong place so you will have to "nibble" > >> out > >> > > the aluminum plate that holds them in the 732A battery pack. You have > >> to > >> > > be > >> > > careful if you use the 12v 5AH as you will have 4 extra battery > >> connection > >> > > leads to deal with and connect correctly. I would stick with the 6V > >> 4AH. > >> > > New batteries will last around 12 to 14 hours before the "CAL" light > >> goes > >> > > out when AC power is not applied. So shipping to Cal Lab can be a > >> problem > >> > > if it is a distance away, or you have to use a shipper like UPS or > >> FEDEX > >> > > and > >> > > you ship the night before and then use their "Morning delivery" and > >> the > >> Cal > >> > > Lab is expecting your 732A. Same on the way back to you. Of course > >> you > >> > > could always strap another battery on the 732A and hook it up to the > >> "ext > >> > > power" plug to last longer. I have seen it done. The issue is to get > >> the > >> > > Cal Lab to charge the extra battery before they ship the 732A back to > >> you. > >> > > > >> > > When you remove and work on the battery pack always have the AC > >> power > >> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated > >> supply > >> > > is working. > >> > > > >> > > The "CAL" light is to indicate that power has not been lost to the > >> > > Reference Amp or other associated circuits. When the raw supply > >> (battery) > >> > > voltage drops below about 21 volts the "CAL" light will go out. Below > >> that > >> > > voltage the heater circuits will not work correctly and the 18.6 volt > >> > > regulated supply will not regulate. The requirement is that the > >> Reference > >> > > Amp be kept "alive" at all times to maintain the output voltage that > >> was > >> > > measured at the time of the most recent Calibration or Certification. > >> When > >> > > the semiconductor junctions are unbiased and cool off when power is > >> lost, > >> > > and then power is restored the result will be a different 10 volts > >> than > >> > > before the power failure. My experience is that after all of the > >> years > >> > > that > >> > > these units have been powered up, this won't happen and when power is > >> lost > >> > > and then restored, even months later, the 732A will come back to > >> almost > >> > > exactly the same 10 volts as when they lost power, usually with in 0.2 > >> PPM > >> > > after 24 hours of "warm up". > >> > > > >> > > What type of hookup leads are you using when measuring the 1 volt > >> > > output? What is the PLC set to? I always use 100 PLC to measure > >> this. > >> If > >> > > you don't have "low thermal" connection leads you can experience uV > >> changes > >> > > for a minute or more after plugging in the leads due to the "thermals" > >> > > generated because of the difference in temperature between the banana > >> jacks > >> > > on the 732A and the banana plugs of the leads. I have found that even > >> just > >> > > plugging in the lead will generate a thermal difference because of > >> > > difference of temps and some heating due to the physical act of just > >> > > inserting the plug because of friction between the jack and plug (my > >> theory > >> > > at any rate). You have to allow at least a minute or more before > >> being > >> > > able > >> > > to make a measurement after plugging in the leads. I just measured > >> the > >> > > variation of the 1 volt output of my 732A and using my 3458A and I > >> got a > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC on > >> the > >> 1 > >> > > volt range of the 3458A. Using the MATH function and all of the data > >> you > >> > > can collect. That was after waiting for several minutes after > >> plugging > >> in > >> > > the leads. > >> > > > >> > > I hope all of this helps. > >> > > > >> > > Bill > >> > > > >> > > > >> > > > >> > > ----- Original Message ----- > >> > > From: "Randy Evans" <randyevans2688@gmail.com> > >> > > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > >> > > Sent: Friday, August 22, 2014 9:03 PM > >> > > Subject: Re: [volt-nuts] 732A and Prologix received > >> > > > >> > > > >> > > > Todd, > >> > > > > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries > >> that > >> I > >> > > > keep topped off and they have very low current draw (~2 to 3 mA at > >> 13.5 > >> > > > VDC) when charged and at their float voltage, so I am pretty sure > >> they > >> > > are > >> > > > in good condition. I will look at getting those in the units after > >> I > >> > > > ascertain the condition of the 732. > >> > > > > >> > > > So now I have a what appears to be a functioning 3458A and a 732A > >> but > >> > > they > >> > > > slightly disagree. I am like the man with two watches that disagree > >> on > >> > > the > >> > > > time - which is correct? For the moment, i am only concerned with > >> > > > stability. The need for absolute accuracy will come later. > >> > > > > >> > > > Randy > >> > > > > >> > > > > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef@gmail.com > >> > > >> > > wrote: > >> > > > > >> > > > > Randy, > >> > > > > > >> > > > > You have two possible choices. It can be configured with 4 x 6v > >> 4Ah > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner > >> has > >> > > > > modified the battery pack already. A couple of mine needed a > >> nibbler > >> > > tool > >> > > > > to remove enough of the aluminum cover that fits over the tops of > >> the > >> > > > > batteries. The original cover will short out to the battery tabs > >> > > regardless > >> > > > > of the battery configuration if this is not done. > >> > > > > > >> > > > > You can find larger capacity batteries that will give you slightly > >> more > >> > > > > battery life. I lost a couple sets of mail-order batteries after a > >> few > >> > > > > extended outages. I would recommend going with locally bought > >> batteries > >> > > > > instead of the cheaper mail order. My local Batteries Plus will > >> > > typically > >> > > > > have some warranty if I remember correctly. Moving forward I will > >> only > >> > > use > >> > > > > 2 12v batteries and pre-charge them on a battery charger to > >> equalize > >> > > them > >> > > > > before putting them in the 732A. I think the cheap batteries did > >> not > >> > > > > discharge equally, and would not recover when power was applied. > >> > > > > > >> > > > > Inspect the back plane for damaged traces and look at the > >> capacitors. I > >> > > had > >> > > > > a few that looked questionable. So far, I have replaced all the > >> big > >> > > caps > >> > > on > >> > > > > the pre-regulator and regulator boards. My feeling is that once > >> these > >> > > go > >> > > > > online, they should run as long as possible between repairs. > >> > > > > > >> > > > > The battery charger circuit may need adjusting. I tweaked mine and > >> it > >> > > > > seemed to work fine. > >> > > > > > >> > > > > Todd > >> > > > > > >> > > > > > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans < > >> > > randyevans2688@gmail.com> > >> > > > > wrote: > >> > > > > > >> > > > > > I received my Fluke 732A today. Just powered it up but it needs > >> new > >> > > > > > batteries. Any suggestions for sources (I haven't opened up the > >> unit > >> > > > > yet - > >> > > > > > I want to make sure it works before doing that). Also received > >> the > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran > >> data > >> > > > > dumper > >> > > > > > program to get the CAL data from my 3458A. Should be a busy > >> weekend. > >> > > > > > > >> > > > > > Randy > >> > > > > > _______________________________________________ > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com > >> > > > > > To unsubscribe, go to > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> > > > > > and follow the instructions there. > >> > > > > > > >> > > > > _______________________________________________ > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com > >> > > > > To unsubscribe, go to > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> > > > > and follow the instructions there. > >> > > > > > >> > > > _______________________________________________ > >> > > > volt-nuts mailing list -- volt-nuts@febo.com > >> > > > To unsubscribe, go to > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> > > > and follow the instructions there. > >> > > > >> > > _______________________________________________ > >> > > volt-nuts mailing list -- volt-nuts@febo.com > >> > > To unsubscribe, go to > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> > > and follow the instructions there. > >> > > > >> > _______________________________________________ > >> > volt-nuts mailing list -- volt-nuts@febo.com > >> > To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> > and follow the instructions there. > >> > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
MS
Mike S
Tue, Aug 26, 2014 3:05 PM

After some more research, I think I've answered some of my own questions -

Tellurium copper is used for binding posts, not because it has any
special thermal or EMF mojo, but because it machines much better than
pure copper. And, I suppose, because it sounds like it's extra special.

The Seebeck coefficients (uV/C, relative to Cu) of some relevant materials:
Cu 0.0
Ag .2
Au .5
Yellow brass 1.5
Phosphor bronze 2.0
63/37 solder 3.0
Sn 3.1
Stainless steel 3.1
Beryllium copper 5.0
Fe -12.3
Ni  22.3
Te -49.25

Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
I'd expect that there might be some coefficient between Cu and CuTe
(0.5% Te), but I could find no reference. The relatively large number
for CuBe is interesting, since that's a common material for banana plug
springs, where one might expect the greatest temperature differential to
occur in such a connection (between the thermal masses of the binding
post/jack and the bulk of the banana plug). Heat has to flow a
considerable distance through the springs, very much more than when it
flows through a surface plating.

The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
contacts. It seems there might be an improvement to be had by using the
older style pin plugs, where a solid pin was partially sliced into 4
sections which were then spread apart a bit to create tension. That
could eliminate relatively large thermocouples at a thermal gradient,
and might also be expected to have less thermal resistance, allowing the
connection to settle quicker.

But maybe not - I'm still not clear on how plated conductors behave in
this situation. For a high impedance voltage measurement where almost no
current flows, the gold plating may carry the signal, so there is no
real thermocouple (or more correctly, it's entirely contained within the
connector). But if that's the case, why fool around with special copper
connectors when common brass ones would be easier/cheaper? For current
or resistance, the signal would also flow through the base metal, so
does this have an effect (especially for tinned copper test leads, where
there may be a larger temperature difference between the ends???

Nickle is avoided as a contact material largely because it is subject to
fretting corrosion. Tests done by AMP
(http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show that
a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a short
time due to this, while Ag and Au plated contacts exhibit negligible
changes.

Cu (with Be for better machining) seems to be used as the base material
for jacks/plugs to get thermal EMF cancellation to the wiring on both
sides (i.e. use copper everywhere except where there is a minimal
thermal gradient, like platings).

--

After some more research, I think I've answered some of my own questions - Tellurium copper is used for binding posts, not because it has any special thermal or EMF mojo, but because it machines much better than pure copper. And, I suppose, because it sounds like it's extra special. The Seebeck coefficients (uV/C, relative to Cu) of some relevant materials: Cu 0.0 Ag .2 Au .5 Yellow brass 1.5 Phosphor bronze 2.0 63/37 solder 3.0 Sn 3.1 Stainless steel 3.1 Beryllium copper 5.0 Fe -12.3 Ni 22.3 Te -49.25 Based on the extreme Seebeck coefficient of pure tellurium vs. copper, I'd expect that there might be some coefficient between Cu and CuTe (0.5% Te), but I could find no reference. The relatively large number for CuBe is interesting, since that's a common material for banana plug springs, where one might expect the greatest temperature differential to occur in such a connection (between the thermal masses of the binding post/jack and the bulk of the banana plug). Heat has to flow a considerable distance through the springs, very much more than when it flows through a surface plating. The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring contacts. It seems there might be an improvement to be had by using the older style pin plugs, where a solid pin was partially sliced into 4 sections which were then spread apart a bit to create tension. That could eliminate relatively large thermocouples at a thermal gradient, and might also be expected to have less thermal resistance, allowing the connection to settle quicker. But maybe not - I'm still not clear on how plated conductors behave in this situation. For a high impedance voltage measurement where almost no current flows, the gold plating may carry the signal, so there is no real thermocouple (or more correctly, it's entirely contained within the connector). But if that's the case, why fool around with special copper connectors when common brass ones would be easier/cheaper? For current or resistance, the signal would also flow through the base metal, so does this have an effect (especially for tinned copper test leads, where there may be a larger temperature difference between the ends??? Nickle is avoided as a contact material largely because it is subject to fretting corrosion. Tests done by AMP (http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show that a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a short time due to this, while Ag and Au plated contacts exhibit negligible changes. Cu (with Be for better machining) seems to be used as the base material for jacks/plugs to get thermal EMF cancellation to the wiring on both sides (i.e. use copper everywhere except where there is a minimal thermal gradient, like platings). --
A
acbern@gmx.de
Tue, Aug 26, 2014 4:56 PM

Dave,

the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements
by E.S.Williams.

Adrian

Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration

Adrian,
Do you have a link or title for the NIST paper that you mentioned?

Dave M

acbern@gmx.de wrote:

fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.

what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be <10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.

adrian

Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration

Well, you sort of answered your own question.  The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple.  Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer.  Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer.  At that point,
the AC voltage source is equal to that of the DC voltage source.

Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard.  They aren't cheap, and you
have to have a converter for each range of voltages that you need to
measure.  The thermal converters used with this type of transfer
standard isn't great (50 MHz or so typical), but their accuracy far
surpasses that of the thermistor type sensors.

There are other brands and models of thermal transfer standards, but
I have a Fluke model 540 and a few thermal converters.  That's why I
referred you to the manual for it.

Cheers,
Dave M

pa4tim@gmail.com wrote:

Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two
AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
ago.

My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for
1V, 10V and 100V (and use several resistors/heaters.)

Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that)  I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.

Fred, pa4tim


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

When the people fear the government, there is tyranny. When the
government fears the people, there is liberty -- Thomas Jefferson

Dave M


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Dave, the title is: Thermal Voltage Converters and Comparator for Very Accurate AC Voltage Measurements by E.S.Williams. Adrian > Gesendet: Montag, 25. August 2014 um 19:32 Uhr > Von: "Dave M" <dgminala@mediacombb.net> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] AC calibration > > Adrian, > Do you have a link or title for the NIST paper that you mentioned? > > Dave M > > > acbern@gmx.de wrote: > > fred, > > generally you raise a good point, I had the same issue of calibrating > > an ac voltage to a high level of accuracy. you need this e.g. to > > validate the self.cal of a 3458a or other precison stuff like the > > 8506a0. > > > > what i would recommend to do if you want to keep costs down is: > > in a nutshell, get a thermal converter in the lowest range you need > > and a second one on range above. build a set of resistor range > > extenders (rf type with appropriate connectors and housings) to > > expand the range to where you need to be max. get one of the thermal > > converter calibrated (the higher one usually, and you need to havr > > good cal lab, should be <10ppm accuracy) and use it to calibrate the > > rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for > > thermal converters! at higher frequencies, due to reflections and > > stray capacitance/inductance influences, the accuracy decreases. the > > resistor range extenders though, if build up correctly, only have a > > few ppm impact (there is a paper from nist on that, but this is only > > typical). you can calibrate all converters to the one you got > > externally calibrated. do some research in the web, when you do the > > calibration, you need to determine the so-called constant N. then do > > an ac, dc+, ac, dc-, ac measurement between the the two and establish > > the deviation, also establish the error propagation. the end result > > will be a set of highly precise (low inaccuracies9 thermal converters > > good enough to calibrate a 3458a an better devices. if you want to > > spend the money, you could also buy a set of converters/range > > resistors (with/without a 540), that typically is a few k altogether, > > while a single device sometimes is available for below 100 bucks. you > > need to have a stable 7.5 digit nanovoltmeter though for the > > measurements of the tvcs (34420a or 2182 typically ) and precision > > (stable) dc and ac sources. but in the end, all you need is a single > > calibrated thermal converter. > > > > adrian > > > > > > > >> Gesendet: Montag, 25. August 2014 um 18:38 Uhr > >> Von: "Dave M" <dgminala@mediacombb.net> > >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > >> Betreff: Re: [volt-nuts] AC calibration > >> > >> Well, you sort of answered your own question. The equipment is > >> called a Thermal Transfer Standard, but instead of thermistors, it > >> uses a thermocouple. Look at the manual for the Fluke 540B > >> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how > >> it's done. Basically, the AC source is input into the transfer > >> standard, and the standard's internal reference voltage is adjusted > >> for a null on the galvanometer. Leaving the reference voltage > >> setting alone, a DC voltage is input into the unit, and the DC > >> source is adjusted for a null on the galvanometer. At that point, > >> the AC voltage source is equal to that of the DC voltage source. > >> > >> Ther are thermocouple-type thermal converters used for RF voltage > >> measurements with the transfer standard. They aren't cheap, and you > >> have to have a converter for each range of voltages that you need to > >> measure. The thermal converters used with this type of transfer > >> standard isn't great (50 MHz or so typical), but their accuracy far > >> surpasses that of the thermistor type sensors. > >> > >> There are other brands and models of thermal transfer standards, but > >> I have a Fluke model 540 and a few thermal converters. That's why I > >> referred you to the manual for it. > >> > >> Cheers, > >> Dave M > >> > >> > >> pa4tim@gmail.com wrote: > >>> Is there a way to link an AC voltage to a DC source for compare. I > >>> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) > >>> against standardcells. But for AC I can not do that. I have two > >>> AC+DC TRMS 7,5 digit meters but the last calibration was 2 years > >>> ago. > >>> > >>> My idea is in theory simple. It is based on the thermal converters > >>> used in RF powermeters. Two resistors, two high resolution > >>> temperature meters. AC on the first en DC on the second. If both are > >>> the same temperature the AC voltage is the same as the DC voltage. > >>> But I'm sure some people here have done this in the past. I would > >>> like to use it for 50 to 100 kHz (or less) and something like for > >>> 1V, 10V and 100V (and use several resistors/heaters.) > >>> > >>> Or mabey there is an other way to convert AC (for RF it can be done > >>> with lightbubs but I never tryed that) I do not mind if it is slow > >>> etc, I like this sort of experiments. You can learn a lot from it. > >>> > >>> Fred, pa4tim > >> > >> > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow > >> the instructions there. > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > When the people fear the government, there is tyranny. When the > government fears the people, there is liberty -- Thomas Jefferson > > > Dave M > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
PK
Poul-Henning Kamp
Tue, Aug 26, 2014 7:42 PM

In message <trinity-9076c25b-a0a9-4b28-b29a-20f9e06039a3-1409072175101@3capp-gm
x-bs56>, acbern@gmx.de writes:

Tellurium/copper is used because it machines much better than pure
copper, (20% -> 90%) without ruining the conductivity too much.

Most other metals which can improve its machinability has big
negative impacts on the conductivity of copper.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- In message <trinity-9076c25b-a0a9-4b28-b29a-20f9e06039a3-1409072175101@3capp-gm x-bs56>, acbern@gmx.de writes: Tellurium/copper is used because it machines much better than pure copper, (20% -> 90%) without ruining the conductivity too much. Most other metals which can improve its machinability has big negative impacts on the conductivity of copper. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
RE
Randy Evans
Tue, Aug 26, 2014 9:16 PM

Bill,

I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.

Randy

On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3637@att.net wrote:

Not that I know of.  Just wait and when you don't see a "SMPL" on the
display it is done.  But then with 1000 PLC that is around 16.66 seconds
per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes.  Not from the front panel
at any rate.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received

Is there any way to tell when the function key routine is complete?  In

the

case of taking multiple readings using the DEFKEY and MATH function, I
don't see any indication when the routine is complete.  In one particular
case, I am taking a 100 readings with NLPC set for 1000 so its a long

while

before it's complete, but i have to guess when it's done.

Thanks,

Randy

On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2688@gmail.com
wrote:

Bill,

I am trying to figure out the MATH function without much success.  I

input

the sequence you said (I looked up the instructions to understand what

you

did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;

and

it shows up on the display when I input BLUE F1.  I hit ENTER and it

takes

the 40 measurements and the MATH symbol shows on the display during the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and

I

get a MATH ERR symbol on the display.  I tried it a couple of times and

the

same result so I am doing something wrong.  Is there a better source

for

explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual

measurements.

Randy

On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3637@att.net wrote:

Randy:

 The MATH function is accessible from the keypad.  I don't have an

IEEE

interface right now that works.  You can also program the numeric

keypad

keys to have preprogrammed functions.  DEFKEY

 I have made my own "low thermal" measurement leads from Pomona

#4892

banana plugs and Belden #9272 wire.  Why 9272, because it was handy at

the

time.  It is tin plated copper, shielded twisted pair 20 ga.  I have

plans

to do custom cables with 16 ga. bare copper wire that I will twist and
then
put a braided shield over it.  I simply cannot find what I want so I

will

build my own cable.  I have done something like this before and it

worked

fine.  When I get a "round toit".

 I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I

have

used
in the past to verify my homemade "low thermal" leads as described

above.

Frankly I cannot see any difference between using the 1756 cables and

my

homemade cables once I give them a few minutes for the thermals to go
away.
As far as I can tell and measure the differences, if any, are below

0.1

ppm
at 10 volts.

 Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are

all

adjustable you may be seeing a misadjusted 1 volt from the 732A.  As

far

as
the instability of the readings it is hard to determine which is

causing

the
problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
following code.  "MATH 14;NRDGS 40;TRIG 4;TRIG ;"  So what this does

is

set
the MATH to "Statistics" (store high reading/low reading/ and mean of

the

readings) in the registers, the number of readings to "40", the

trigger
to

"hold" (which keeps the meter from triggering until I press "ENTER"

and

then
trigger the sequence of 40 readings when I push the "ENTER" button.

You

can
do all of this manually from the keypads but since I use this sequence

a

lot
I have preprogrammed it.  This is after I set digits to "8" and PLC to
100.
Once those 40 readings are finished then you can access the various

MATH

statistic registers, using the menu, by entering "MATH" and then a 2

for

low, a 4 for mean, and 13 for high.  Of course you could do all of

this

through the IEEE also.  The 3458A has a very rich set of measurement
commands.  I am still learning all of them.  It depends upon what I am
trying to accomplish.

 Since the 1.018 and 1.0 volt outputs are passive and derived from

resistive dividers from the 10 volt, I don't see how they could

contribute

to the varying readings you are measuring.  I think I would put a

short
on

the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved.  When I do this I

see a

variation from low reading to high reading of 0.125 uVolts and then
another
40 I get 0.155 uVolts.  This is without the GUARD connected to the low
side
of the measurment terminals, GUARD connected doesn't seem to affect

the

readings.  So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output

I

got
a variation of 0.159 uVolts using the same 40 reading method above.  I
would
use this to determine where your problem might exist.  Just having the
meter
input shorted will point you in the right direction.  Meter, cables or
732A.

 Sorry for the long dissertation.  Friends get mad at me for being

so

detailed sometimes.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received

Bill,

I have convinced myself that the problem I an seeing is due to

thermals.

If I move the cables (with gold-plated banana plugs) using a small

towel

rather than letting my hand touch the plugs, it is much more stable.

If I

then hold the banana plug with my hand after the reading has

stabilized,

the reading drifts rapidly upward.  I am trying to check the

stability

of

the reading but I haven't figured out the MATH function yet.  I

assume

this

is a programmed function using GPIB only?

The stability I am seeing by manually recording readings (using NLPC

of

100

and 1000) is much greater than what you are measuring on your

system.

Not

sure how to ascertain if it's the 3458A or the 732.  The value of

the

readings are very different between the two - the 3458 reads about

50
uV

high on the 10 V output and about 12 uV low on the 1V output.

Rather

large

differences (this is after an ACAL).  I need to find some better

cables

to

make sure the errors are not due to thermals again.

Randy

On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3637@att.net

wrote:

Randy:

 I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will

fit

perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C

and

others
that are in this size and package.  Order from one of the usual

electronics

distributors like Allied, Mouser, Digikey.  This is a very common

battery

as
it is used in a lot of "EXIT" signs so they are lighted when the

power

goes

out.  I don't see how the 12V 7AH will fit as they are too large.

I

guess

you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6

volt

4
AH

but the terminals are in the wrong place so you will have to

"nibble"

out

the aluminum plate that holds them in the 732A battery pack.  You

have

to

be
careful if you use the 12v 5AH as you will have 4 extra battery

connection

leads to deal with and connect correctly.  I would stick with the

6V

4AH.

New batteries will last around 12 to 14 hours before the "CAL"

light

goes

out when AC power is not applied.  So shipping to Cal Lab can be a

problem

if it is a distance away, or you have to use a shipper like UPS or

FEDEX

and
you ship the night before and then use their "Morning delivery"

and

the
Cal

Lab is expecting your 732A.  Same on the way back to you.  Of

course

you

could always strap another battery on the 732A and hook it up to

the

"ext

power" plug to last longer.  I have seen it done.  The issue is to

get

the

Cal Lab to charge the extra battery before they ship the 732A back

to

you.

 When you remove and work on the battery pack always have the

AC

power

plugged in.  The "CAL" led will stay on because the 18.6 v

regulated

supply

is working.

 The "CAL" light is to indicate that power has not been lost to

the

Reference Amp or other associated circuits.  When the raw supply

(battery)

voltage drops below about 21 volts the "CAL" light will go out.

Below

that

voltage the heater circuits will not work correctly and the 18.6

volt

regulated supply will not regulate.  The requirement is that the

Reference

Amp be kept "alive" at all times to maintain the output voltage

that

was

measured at the time of the most recent Calibration or

Certification.

When

the semiconductor junctions are unbiased and cool off when power

is

lost,

and then power is restored the result will be a different 10 volts

than

before the power failure.  My experience is that after all of the

years

that
these units have been powered up, this won't happen and when power

is

lost

and then restored, even months later, the 732A will come back to

almost

exactly the same 10 volts as when they lost power, usually with in

0.2

PPM

after 24 hours of "warm up".

 What type of hookup leads are you using when measuring the 1

volt

output?  What is the PLC set to?  I always use 100 PLC to measure

this.
If

you don't have "low thermal" connection leads you can experience

uV

changes

for a minute or more after plugging in the leads due to the

"thermals"

generated because of the difference in temperature between the

banana

jacks

on the 732A and the banana plugs of the leads.  I have found that

even

just

plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of

just

inserting the plug because of friction between the jack and plug

(my

theory

at any rate).  You have to allow at least a minute or more before

being

able
to make a measurement after plugging in the leads.  I just

measured

the

variation of the 1 volt output of my 732A and using my 3458A and I

got a

total difference of  0.159 uV over 40 measurements using 100 PLC

on

the
1

volt range of the 3458A.  Using the MATH function and all of the

data

you

can collect.  That was after waiting for several minutes after

plugging
in

the leads.

 I hope all of this helps.

Bill

----- Original Message -----
From: "Randy Evans" randyevans2688@gmail.com
To: "Discussion of precise voltage measurement"

Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received

Todd,

Thanks for the info.  I have several Panasonic 12V 7 AH

batteries

that
I

keep topped off and they have very low current draw (~2 to 3 mA

at

13.5

VDC) when charged and at their float voltage, so I am pretty

sure

they

are

in good condition.  I will look at getting those in the units

after

I

ascertain the condition of the 732.

So now I have a what appears to be a functioning 3458A and a

732A

but

they

slightly disagree.  I am like the man with two watches that

disagree

on

the

time  - which is correct?  For the moment, i am only concerned

with

stability.  The need for absolute accuracy will come later.

Randy

On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef

wrote:

Randy,

You have two possible choices. It can be configured with 4 x

6v

4Ah

batteries or 2 x 12v 7Ah batteries. Hopefully the previous

owner

has

modified the battery pack already. A couple of mine needed a

nibbler

tool

to remove enough of the aluminum cover that fits over the tops

of

the

batteries. The original cover will short out to the battery

tabs

regardless

of the battery configuration if this is not done.

You can find larger capacity batteries that will give you

slightly

more

battery life. I lost a couple sets of mail-order batteries

after a

few

extended outages. I would recommend going with locally bought

batteries

instead of the cheaper mail order. My local Batteries Plus

will

typically

have some warranty if I remember correctly. Moving forward I

will

only

use

2 12v batteries and pre-charge them on a battery charger to

equalize

them

before putting them in the 732A. I think the cheap batteries

did

not

discharge equally, and would not recover when power was

applied.

Inspect the back plane for damaged traces and look at the

capacitors. I

had

a few that looked questionable. So far, I have replaced all

the

big

caps
on

the pre-regulator and regulator boards. My feeling is that

once

these

go

online, they should run as long as possible between repairs.

The battery charger circuit may need adjusting. I tweaked mine

and

it

seemed to work fine.

Todd

On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <

wrote:

I received my Fluke 732A today.  Just powered it up but it

needs

new

batteries.  Any suggestions for sources (I haven't opened up

the

unit

yet -

I want to make sure it works before doing that).  Also

received

the

ProLogix USB-GPIB adapter.  I plan on using Mark Sims' CAL

ran

data

dumper

program to get the CAL data from my 3458A.  Should be a busy

weekend.

Randy


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Bill, I measured the time between SMPL symbols with NPLC set to 1000 and it is approximately 33 seconds. It takes an hour to complete 100 readings. Randy On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold <wpgold3637@att.net> wrote: > Not that I know of. Just wait and when you don't see a "SMPL" on the > display it is done. But then with 1000 PLC that is around 16.66 seconds > per > reading times 100 readings is somewhere around 28 minutes and there is > probably some overhead time so around 30 minutes. Not from the front panel > at any rate. > > Bill > > ----- Original Message ----- > From: "Randy Evans" <randyevans2688@gmail.com> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Sent: Monday, August 25, 2014 9:23 PM > Subject: Re: [volt-nuts] 732A and Prologix received > > > > Is there any way to tell when the function key routine is complete? In > the > > case of taking multiple readings using the DEFKEY and MATH function, I > > don't see any indication when the routine is complete. In one particular > > case, I am taking a 100 readings with NLPC set for 1000 so its a long > while > > before it's complete, but i have to guess when it's done. > > > > Thanks, > > > > Randy > > > > > > On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans <randyevans2688@gmail.com> > > wrote: > > > > > Bill, > > > > > > I am trying to figure out the MATH function without much success. I > input > > > the sequence you said (I looked up the instructions to understand what > you > > > did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; > and > > > it shows up on the display when I input BLUE F1. I hit ENTER and it > takes > > > the 40 measurements and the MATH symbol shows on the display during the > > > measurements. After the SMPL symbol no longer blinks I hit MATH 2 and > I > > > get a MATH ERR symbol on the display. I tried it a couple of times and > the > > > same result so I am doing something wrong. Is there a better source > for > > > explaining how to do front panel masurements than the User Guide, which > > > seems oriented at programming automatic rather than manual > measurements. > > > > > > Randy > > > > > > > > > > > > On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637@att.net> wrote: > > > > > >> Randy: > > >> > > >> The MATH function is accessible from the keypad. I don't have an > IEEE > > >> interface right now that works. You can also program the numeric > keypad > > >> keys to have preprogrammed functions. DEFKEY > > >> > > >> I have made my own "low thermal" measurement leads from Pomona > #4892 > > >> banana plugs and Belden #9272 wire. Why 9272, because it was handy at > the > > >> time. It is tin plated copper, shielded twisted pair 20 ga. I have > plans > > >> to do custom cables with 16 ga. bare copper wire that I will twist and > > >> then > > >> put a braided shield over it. I simply cannot find what I want so I > will > > >> build my own cable. I have done something like this before and it > worked > > >> fine. When I get a "round toit". > > >> > > >> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I > have > > >> used > > >> in the past to verify my homemade "low thermal" leads as described > above. > > >> Frankly I cannot see any difference between using the 1756 cables and > my > > >> homemade cables once I give them a few minutes for the thermals to go > > >> away. > > >> As far as I can tell and measure the differences, if any, are below > 0.1 > > >> ppm > > >> at 10 volts. > > >> > > >> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are > all > > >> adjustable you may be seeing a misadjusted 1 volt from the 732A. As > far > > >> as > > >> the instability of the readings it is hard to determine which is > causing > > >> the > > >> problem. I have programed (DEFKEY) a numeric keypad key #1 with the > > >> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does > is > > >> set > > >> the MATH to "Statistics" (store high reading/low reading/ and mean of > the > > >> readings) in the registers, the number of readings to "40", the > trigger > to > > >> "hold" (which keeps the meter from triggering until I press "ENTER" > and > > >> then > > >> trigger the sequence of 40 readings when I push the "ENTER" button. > You > > >> can > > >> do all of this manually from the keypads but since I use this sequence > a > > >> lot > > >> I have preprogrammed it. This is after I set digits to "8" and PLC to > > >> 100. > > >> Once those 40 readings are finished then you can access the various > MATH > > >> statistic registers, using the menu, by entering "MATH" and then a 2 > for > > >> low, a 4 for mean, and 13 for high. Of course you could do all of > this > > >> through the IEEE also. The 3458A has a very rich set of measurement > > >> commands. I am still learning all of them. It depends upon what I am > > >> trying to accomplish. > > >> > > >> Since the 1.018 and 1.0 volt outputs are passive and derived from > > >> resistive dividers from the 10 volt, I don't see how they could > contribute > > >> to the varying readings you are measuring. I think I would put a > short > on > > >> the input of the 3458A and manually set the range to 1 volt and then > > >> observe > > >> the variations that way without the 732A involved. When I do this I > see a > > >> variation from low reading to high reading of 0.125 uVolts and then > > >> another > > >> 40 I get 0.155 uVolts. This is without the GUARD connected to the low > > >> side > > >> of the measurment terminals, GUARD connected doesn't seem to affect > the > > >> readings. So that is the base noise of the 3458A without the 732A, > > >> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output > I > > >> got > > >> a variation of 0.159 uVolts using the same 40 reading method above. I > > >> would > > >> use this to determine where your problem might exist. Just having the > > >> meter > > >> input shorted will point you in the right direction. Meter, cables or > > >> 732A. > > >> > > >> Sorry for the long dissertation. Friends get mad at me for being > so > > >> detailed sometimes. > > >> > > >> Bill > > >> > > >> ----- Original Message ----- > > >> From: "Randy Evans" <randyevans2688@gmail.com> > > >> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > >> Sent: Sunday, August 24, 2014 7:22 AM > > >> Subject: Re: [volt-nuts] 732A and Prologix received > > >> > > >> > > >> > Bill, > > >> > > > >> > I have convinced myself that the problem I an seeing is due to > thermals. > > >> > If I move the cables (with gold-plated banana plugs) using a small > > >> towel > > >> > rather than letting my hand touch the plugs, it is much more stable. > > >> If I > > >> > then hold the banana plug with my hand after the reading has > stabilized, > > >> > the reading drifts rapidly upward. I am trying to check the > stability > > >> of > > >> > the reading but I haven't figured out the MATH function yet. I > assume > > >> this > > >> > is a programmed function using GPIB only? > > >> > > > >> > The stability I am seeing by manually recording readings (using NLPC > of > > >> 100 > > >> > and 1000) is much greater than what you are measuring on your > system. > > >> Not > > >> > sure how to ascertain if it's the 3458A or the 732. The value of > the > > >> > readings are very different between the two - the 3458 reads about > 50 > uV > > >> > high on the 10 V output and about 12 uV low on the 1V output. > Rather > > >> large > > >> > differences (this is after an ACAL). I need to find some better > cables > > >> to > > >> > make sure the errors are not due to thermals again. > > >> > > > >> > Randy > > >> > > > >> > > > >> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637@att.net> > wrote: > > >> > > > >> > > Randy: > > >> > > > > >> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will > fit > > >> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C > and > > >> > > others > > >> > > that are in this size and package. Order from one of the usual > > >> electronics > > >> > > distributors like Allied, Mouser, Digikey. This is a very common > > >> battery > > >> > > as > > >> > > it is used in a lot of "EXIT" signs so they are lighted when the > power > > >> goes > > >> > > out. I don't see how the 12V 7AH will fit as they are too large. > I > > >> guess > > >> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 > volt > > >> 4 > > >> AH > > >> > > but the terminals are in the wrong place so you will have to > "nibble" > > >> out > > >> > > the aluminum plate that holds them in the 732A battery pack. You > have > > >> to > > >> > > be > > >> > > careful if you use the 12v 5AH as you will have 4 extra battery > > >> connection > > >> > > leads to deal with and connect correctly. I would stick with the > 6V > > >> 4AH. > > >> > > New batteries will last around 12 to 14 hours before the "CAL" > light > > >> goes > > >> > > out when AC power is not applied. So shipping to Cal Lab can be a > > >> problem > > >> > > if it is a distance away, or you have to use a shipper like UPS or > > >> FEDEX > > >> > > and > > >> > > you ship the night before and then use their "Morning delivery" > and > > >> the > > >> Cal > > >> > > Lab is expecting your 732A. Same on the way back to you. Of > course > > >> you > > >> > > could always strap another battery on the 732A and hook it up to > the > > >> "ext > > >> > > power" plug to last longer. I have seen it done. The issue is to > get > > >> the > > >> > > Cal Lab to charge the extra battery before they ship the 732A back > to > > >> you. > > >> > > > > >> > > When you remove and work on the battery pack always have the > AC > > >> power > > >> > > plugged in. The "CAL" led will stay on because the 18.6 v > regulated > > >> supply > > >> > > is working. > > >> > > > > >> > > The "CAL" light is to indicate that power has not been lost to > the > > >> > > Reference Amp or other associated circuits. When the raw supply > > >> (battery) > > >> > > voltage drops below about 21 volts the "CAL" light will go out. > Below > > >> that > > >> > > voltage the heater circuits will not work correctly and the 18.6 > volt > > >> > > regulated supply will not regulate. The requirement is that the > > >> Reference > > >> > > Amp be kept "alive" at all times to maintain the output voltage > that > > >> was > > >> > > measured at the time of the most recent Calibration or > Certification. > > >> When > > >> > > the semiconductor junctions are unbiased and cool off when power > is > > >> lost, > > >> > > and then power is restored the result will be a different 10 volts > > >> than > > >> > > before the power failure. My experience is that after all of the > > >> years > > >> > > that > > >> > > these units have been powered up, this won't happen and when power > is > > >> lost > > >> > > and then restored, even months later, the 732A will come back to > > >> almost > > >> > > exactly the same 10 volts as when they lost power, usually with in > 0.2 > > >> PPM > > >> > > after 24 hours of "warm up". > > >> > > > > >> > > What type of hookup leads are you using when measuring the 1 > volt > > >> > > output? What is the PLC set to? I always use 100 PLC to measure > > >> this. > > >> If > > >> > > you don't have "low thermal" connection leads you can experience > uV > > >> changes > > >> > > for a minute or more after plugging in the leads due to the > "thermals" > > >> > > generated because of the difference in temperature between the > banana > > >> jacks > > >> > > on the 732A and the banana plugs of the leads. I have found that > even > > >> just > > >> > > plugging in the lead will generate a thermal difference because of > > >> > > difference of temps and some heating due to the physical act of > just > > >> > > inserting the plug because of friction between the jack and plug > (my > > >> theory > > >> > > at any rate). You have to allow at least a minute or more before > > >> being > > >> > > able > > >> > > to make a measurement after plugging in the leads. I just > measured > > >> the > > >> > > variation of the 1 volt output of my 732A and using my 3458A and I > > >> got a > > >> > > total difference of 0.159 uV over 40 measurements using 100 PLC > on > > >> the > > >> 1 > > >> > > volt range of the 3458A. Using the MATH function and all of the > data > > >> you > > >> > > can collect. That was after waiting for several minutes after > > >> plugging > > >> in > > >> > > the leads. > > >> > > > > >> > > I hope all of this helps. > > >> > > > > >> > > Bill > > >> > > > > >> > > > > >> > > > > >> > > ----- Original Message ----- > > >> > > From: "Randy Evans" <randyevans2688@gmail.com> > > >> > > To: "Discussion of precise voltage measurement" > <volt-nuts@febo.com> > > >> > > Sent: Friday, August 22, 2014 9:03 PM > > >> > > Subject: Re: [volt-nuts] 732A and Prologix received > > >> > > > > >> > > > > >> > > > Todd, > > >> > > > > > >> > > > Thanks for the info. I have several Panasonic 12V 7 AH > batteries > > >> that > > >> I > > >> > > > keep topped off and they have very low current draw (~2 to 3 mA > at > > >> 13.5 > > >> > > > VDC) when charged and at their float voltage, so I am pretty > sure > > >> they > > >> > > are > > >> > > > in good condition. I will look at getting those in the units > after > > >> I > > >> > > > ascertain the condition of the 732. > > >> > > > > > >> > > > So now I have a what appears to be a functioning 3458A and a > 732A > > >> but > > >> > > they > > >> > > > slightly disagree. I am like the man with two watches that > disagree > > >> on > > >> > > the > > >> > > > time - which is correct? For the moment, i am only concerned > with > > >> > > > stability. The need for absolute accuracy will come later. > > >> > > > > > >> > > > Randy > > >> > > > > > >> > > > > > >> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef > <tmicallef@gmail.com > > >> > > > >> > > wrote: > > >> > > > > > >> > > > > Randy, > > >> > > > > > > >> > > > > You have two possible choices. It can be configured with 4 x > 6v > > >> 4Ah > > >> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous > owner > > >> has > > >> > > > > modified the battery pack already. A couple of mine needed a > > >> nibbler > > >> > > tool > > >> > > > > to remove enough of the aluminum cover that fits over the tops > of > > >> the > > >> > > > > batteries. The original cover will short out to the battery > tabs > > >> > > regardless > > >> > > > > of the battery configuration if this is not done. > > >> > > > > > > >> > > > > You can find larger capacity batteries that will give you > slightly > > >> more > > >> > > > > battery life. I lost a couple sets of mail-order batteries > after a > > >> few > > >> > > > > extended outages. I would recommend going with locally bought > > >> batteries > > >> > > > > instead of the cheaper mail order. My local Batteries Plus > will > > >> > > typically > > >> > > > > have some warranty if I remember correctly. Moving forward I > will > > >> only > > >> > > use > > >> > > > > 2 12v batteries and pre-charge them on a battery charger to > > >> equalize > > >> > > them > > >> > > > > before putting them in the 732A. I think the cheap batteries > did > > >> not > > >> > > > > discharge equally, and would not recover when power was > applied. > > >> > > > > > > >> > > > > Inspect the back plane for damaged traces and look at the > > >> capacitors. I > > >> > > had > > >> > > > > a few that looked questionable. So far, I have replaced all > the > > >> big > > >> > > caps > > >> > > on > > >> > > > > the pre-regulator and regulator boards. My feeling is that > once > > >> these > > >> > > go > > >> > > > > online, they should run as long as possible between repairs. > > >> > > > > > > >> > > > > The battery charger circuit may need adjusting. I tweaked mine > and > > >> it > > >> > > > > seemed to work fine. > > >> > > > > > > >> > > > > Todd > > >> > > > > > > >> > > > > > > >> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans < > > >> > > randyevans2688@gmail.com> > > >> > > > > wrote: > > >> > > > > > > >> > > > > > I received my Fluke 732A today. Just powered it up but it > needs > > >> new > > >> > > > > > batteries. Any suggestions for sources (I haven't opened up > the > > >> unit > > >> > > > > yet - > > >> > > > > > I want to make sure it works before doing that). Also > received > > >> the > > >> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL > ran > > >> data > > >> > > > > dumper > > >> > > > > > program to get the CAL data from my 3458A. Should be a busy > > >> weekend. > > >> > > > > > > > >> > > > > > Randy > > >> > > > > > _______________________________________________ > > >> > > > > > volt-nuts mailing list -- volt-nuts@febo.com > > >> > > > > > To unsubscribe, go to > > >> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > >> > > > > > and follow the instructions there. > > >> > > > > > > > >> > > > > _______________________________________________ > > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com > > >> > > > > To unsubscribe, go to > > >> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > >> > > > > and follow the instructions there. > > >> > > > > > > >> > > > _______________________________________________ > > >> > > > volt-nuts mailing list -- volt-nuts@febo.com > > >> > > > To unsubscribe, go to > > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > >> > > > and follow the instructions there. > > >> > > > > >> > > _______________________________________________ > > >> > > volt-nuts mailing list -- volt-nuts@febo.com > > >> > > To unsubscribe, go to > > >> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > >> > > and follow the instructions there. > > >> > > > > >> > _______________________________________________ > > >> > volt-nuts mailing list -- volt-nuts@febo.com > > >> > To unsubscribe, go to > > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > >> > and follow the instructions there. > > >> > > >> _______________________________________________ > > >> volt-nuts mailing list -- volt-nuts@febo.com > > >> To unsubscribe, go to > > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > >> and follow the instructions there. > > >> > > > > > > > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
DM
Dave M
Wed, Aug 27, 2014 12:01 AM

Great!!@! Many Thanks!!

Dave M

acbern@gmx.de wrote:

Dave,

the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC
Voltage Measurements by E.S.Williams.

Adrian

Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration

Adrian,
Do you have a link or title for the NIST paper that you mentioned?

Great!!@! Many Thanks!! Dave M acbern@gmx.de wrote: > Dave, > > the title is: > Thermal Voltage Converters and Comparator for Very Accurate AC > Voltage Measurements by E.S.Williams. > > Adrian > > >> Gesendet: Montag, 25. August 2014 um 19:32 Uhr >> Von: "Dave M" <dgminala@mediacombb.net> >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> Betreff: Re: [volt-nuts] AC calibration >> >> Adrian, >> Do you have a link or title for the NIST paper that you mentioned? >