Discussion of precise voltage measurement

View all threads

732A and Prologix received

D
Don@True-Cal
Sat, Aug 30, 2014 9:14 PM

Woops sorry, meant to type "don't use an NPLC setting over 300 (5-sec)" not 5-min. Actually, with AZERO=on, I guess that's 5 seconds of integration over 10 seconds or time.

Don

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the same except I only do voltage so far. I use NPLC 1000 which looks smoother but doesn't really have any additional useful info. I have been meaning to have my TEMP? servoed  some time in the near future. My TEMP?
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie

On 8/29/2014 1:36 PM, Charles Black wrote:


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Woops sorry, meant to type "don't use an NPLC setting over 300 (5-sec)" not 5-min. Actually, with AZERO=on, I guess that's 5 seconds of integration over 10 seconds or time. Don -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Black Sent: Friday, August 29, 2014 3:44 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 732A and Prologix received Hi Don, I'm a bit of trouble sending you this message for some reason. Maybe the third ti me is the charm. Thanks for sharing the graph a few days ago. My 3458A graphs look about the same except I only do voltage so far. I use NPLC 1000 which looks smoother but doesn't really have any additional useful info. I have been meaning to have my TEMP? servoed some time in the near future. My TEMP? runs about 37C at a room temp of 23C. Is your room temp a bit higher? Charlie On 8/29/2014 1:36 PM, Charles Black wrote: > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
D
Don@True-Cal
Sun, Aug 31, 2014 2:47 PM

Charlie,

Did my reply to your question along with my lab temperature graph link come through? Just checking as there has been some issues with the posts.

I don't like the "feature" where you don't see your own posts.

Don

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the same except I only do voltage so far. I use NPLC 1000 which looks smoother but doesn't really have any additional useful info. I have been meaning to have my TEMP? servoed  some time in the near future. My TEMP?
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie

On 8/29/2014 1:36 PM, Charles Black wrote:


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Charlie, Did my reply to your question along with my lab temperature graph link come through? Just checking as there has been some issues with the posts. I don't like the "feature" where you don't see your own posts. Don -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Black Sent: Friday, August 29, 2014 3:44 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 732A and Prologix received Hi Don, I'm a bit of trouble sending you this message for some reason. Maybe the third ti me is the charm. Thanks for sharing the graph a few days ago. My 3458A graphs look about the same except I only do voltage so far. I use NPLC 1000 which looks smoother but doesn't really have any additional useful info. I have been meaning to have my TEMP? servoed some time in the near future. My TEMP? runs about 37C at a room temp of 23C. Is your room temp a bit higher? Charlie On 8/29/2014 1:36 PM, Charles Black wrote: > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
CB
Charles Black
Sun, Aug 31, 2014 5:12 PM

Hi Don,

Your post and link are fine but I need to get some time today to look at
it more carefully.

I had two posts recently that were blank but since then all seems back
to normal.

Charlie

On 8/31/2014 7:47 AM, Don@True-Cal wrote:

Charlie,

Did my reply to your question along with my lab temperature graph link come through? Just checking as there has been some issues with the posts.

I don't like the "feature" where you don't see your own posts.

Don

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the same except I only do voltage so far. I use NPLC 1000 which looks smoother but doesn't really have any additional useful info. I have been meaning to have my TEMP? servoed  some time in the near future. My TEMP?
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie

On 8/29/2014 1:36 PM, Charles Black wrote:


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Don, Your post and link are fine but I need to get some time today to look at it more carefully. I had two posts recently that were blank but since then all seems back to normal. Charlie On 8/31/2014 7:47 AM, Don@True-Cal wrote: > Charlie, > > Did my reply to your question along with my lab temperature graph link come through? Just checking as there has been some issues with the posts. > > I don't like the "feature" where you don't see your own posts. > > Don > > -----Original Message----- > From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Charles Black > Sent: Friday, August 29, 2014 3:44 PM > To: Discussion of precise voltage measurement > Subject: Re: [volt-nuts] 732A and Prologix received > > Hi Don, > > I'm a bit of trouble sending you this message for some reason. Maybe the third ti me is the charm. > > Thanks for sharing the graph a few days ago. My 3458A graphs look about the same except I only do voltage so far. I use NPLC 1000 which looks smoother but doesn't really have any additional useful info. I have been meaning to have my TEMP? servoed some time in the near future. My TEMP? > runs about 37C at a room temp of 23C. Is your room temp a bit higher? > > Charlie > > > On 8/29/2014 1:36 PM, Charles Black wrote: >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> >> >> > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > >