Going from memory, no, you have to pull the roms and read them. At
least, that's what a few of us in the group have done.
You can access ram and nvram, though.
They said 'Windows or better'
so I used Linux.
Actually, I believe that the MONITOR command will allow you look at storage anywhere within the address space of either of the processors, see pages 8.1 through 8.5 of the service manual. So it should be possible to READ the ROMs this way if both processors are working and communicating.
However it would be pretty slow, and as all the ROMs are socketed, it's not a big deal to just remove them and read them in a programmer.
Note that I haven't tried this.
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell
Sent: 13 January 2013 13:28
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Going from memory, no, you have to pull the roms and read them. At least, that's what a few of us in the group have done.
You can access ram and nvram, though.
They said 'Windows or better'
so I used Linux.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi David,
well it is currently working and close to calibrated as far as I can tell. But I want to get those calibration factors saved before it dies on me. Last calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing else to confirm anything else as I have only just got it chatting out through the rs232.
I will have to buy something to read them with.
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 16:36:52 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Actually, I believe that the MONITOR command will allow you look at storage anywhere within the address space of either of the processors, see pages 8.1 through 8.5 of the service manual. So it should be possible to READ the ROMs this way if both processors are working and communicating.
However it would be pretty slow, and as all the ROMs are socketed, it's not a big deal to just remove them and read them in a programmer.
Note that I haven't tried this.
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell
Sent: 13 January 2013 13:28
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Going from memory, no, you have to pull the roms and read them. At least, that's what a few of us in the group have done.
You can access ram and nvram, though.
They said 'Windows or better'
so I used Linux.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
If all you want is to save the calibration factors, just issue CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH - barring component failure that should be good for another ten years or so. You can also display the NVM contents with the MONITOR command, but NOT re-write them.
The only way that I am aware of that will write the caibration constants if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell. But I want to get those calibration factors saved before it dies on me. Last calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing else to confirm anything else as I have only just got it chatting out through the rs232.
I will have to buy something to read them with.
M K
I just noticed a gotcha, I need to set the key to calibrate before I may issue those commands, I guess I need a key. I have printed off the photo and measurements someone on this board did, now to get to a key shop...
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 17:14:07 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
If all you want is to save the calibration factors, just issue CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH - barring component failure that should be good for another ten years or so. You can also display the NVM contents with the MONITOR command, but NOT re-write them.
The only way that I am aware of that will write the caibration constants if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell. But I want to get those calibration factors saved before it dies on me. Last calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing else to confirm anything else as I have only just got it chatting out through the rs232.
I will have to buy something to read them with.
M K
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
MK,
Where are you? If you can't get a key made, I can probably send you one.
IIRC, you can 'activate' the key by removing a wire from the key switch.
That places it in 'CAL' mode. The schematic easily demonstrates this. I
don't think it was 'shorting' the two switch wires. Of course that requires
you to remove the top of the meter.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Tuesday, January 15, 2013 4:30 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
I just noticed a gotcha, I need to set the key to calibrate before I may
issue those commands, I guess I need a key. I have printed off the photo and
measurements someone on this board did, now to get to a key shop...
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 17:14:07 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
If all you want is to save the calibration factors, just issue
CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH -
barring component failure that should be good for another ten years or so.
You can also display the NVM contents with the MONITOR command, but NOT
re-write them.
The only way that I am aware of that will write the caibration constants
if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just
ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]
On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell.
But I want to get those calibration factors saved before it dies on me. Last
calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on
when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing
else to confirm anything else as I have only just got it chatting out
through the rs232.
I will have to buy something to read them with.
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
MK,
All you need to do is unplug the jones plug where the key wiring goes to
the circuit board. No need to do any wiring!
Charlie
On 1/15/2013 5:00 PM, J. L. Trantham wrote:
MK,
Where are you? If you can't get a key made, I can probably send you one.
IIRC, you can 'activate' the key by removing a wire from the key switch.
That places it in 'CAL' mode. The schematic easily demonstrates this. I
don't think it was 'shorting' the two switch wires. Of course that requires
you to remove the top of the meter.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Tuesday, January 15, 2013 4:30 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
I just noticed a gotcha, I need to set the key to calibrate before I may
issue those commands, I guess I need a key. I have printed off the photo and
measurements someone on this board did, now to get to a key shop...
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 17:14:07 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
If all you want is to save the calibration factors, just issue
CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH -
barring component failure that should be good for another ten years or so.
You can also display the NVM contents with the MONITOR command, but NOT
re-write them.
The only way that I am aware of that will write the caibration constants
if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just
ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]
On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell.
But I want to get those calibration factors saved before it dies on me. Last
calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on
when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing
else to confirm anything else as I have only just got it chatting out
through the rs232.
I will have to buy something to read them with.
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Thanks for that, I thought someone had once said that would work, but until I know it better I have been trying to change it as little as possible.
Regards,
M K
Date: Tue, 15 Jan 2013 18:08:02 -0800
From: cblack@centurytel.net
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
MK,
All you need to do is unplug the jones plug where the key wiring goes to
the circuit board. No need to do any wiring!
Charlie
On 1/15/2013 5:00 PM, J. L. Trantham wrote:
MK,
Where are you? If you can't get a key made, I can probably send you one.
IIRC, you can 'activate' the key by removing a wire from the key switch.
That places it in 'CAL' mode. The schematic easily demonstrates this. I
don't think it was 'shorting' the two switch wires. Of course that requires
you to remove the top of the meter.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Tuesday, January 15, 2013 4:30 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
I just noticed a gotcha, I need to set the key to calibrate before I may
issue those commands, I guess I need a key. I have printed off the photo and
measurements someone on this board did, now to get to a key shop...
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 17:14:07 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
If all you want is to save the calibration factors, just issue
CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH -
barring component failure that should be good for another ten years or so.
You can also display the NVM contents with the MONITOR command, but NOT
re-write them.
The only way that I am aware of that will write the caibration constants
if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just
ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]
On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell.
But I want to get those calibration factors saved before it dies on me. Last
calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on
when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing
else to confirm anything else as I have only just got it chatting out
through the rs232.
I will have to buy something to read them with.
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi Joe,
I am in the UK near Guildford. But give me time to get to the locksmith at the weekend first.
And I am going to try just unplugging the cal key too.
Regards,
M K
From: jltran@att.net
To: volt-nuts@febo.com
Date: Tue, 15 Jan 2013 19:00:57 -0600
Subject: Re: [volt-nuts] vSolartron7081 rom reading
MK,
Where are you? If you can't get a key made, I can probably send you one.
IIRC, you can 'activate' the key by removing a wire from the key switch.
That places it in 'CAL' mode. The schematic easily demonstrates this. I
don't think it was 'shorting' the two switch wires. Of course that requires
you to remove the top of the meter.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Tuesday, January 15, 2013 4:30 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
I just noticed a gotcha, I need to set the key to calibrate before I may
issue those commands, I guess I need a key. I have printed off the photo and
measurements someone on this board did, now to get to a key shop...
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 17:14:07 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
If all you want is to save the calibration factors, just issue
CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH -
barring component failure that should be good for another ten years or so.
You can also display the NVM contents with the MONITOR command, but NOT
re-write them.
The only way that I am aware of that will write the caibration constants
if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just
ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]
On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell.
But I want to get those calibration factors saved before it dies on me. Last
calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on
when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing
else to confirm anything else as I have only just got it chatting out
through the rs232.
I will have to buy something to read them with.
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi Charlie,
partial success, calibrate,refresh worked, the display told me so, but the calibrate,dump command did not do anything visible.
M K
Date: Tue, 15 Jan 2013 18:08:02 -0800
From: cblack@centurytel.net
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
MK,
All you need to do is unplug the jones plug where the key wiring goes to
the circuit board. No need to do any wiring!
Charlie
On 1/15/2013 5:00 PM, J. L. Trantham wrote:
MK,
Where are you? If you can't get a key made, I can probably send you one.
IIRC, you can 'activate' the key by removing a wire from the key switch.
That places it in 'CAL' mode. The schematic easily demonstrates this. I
don't think it was 'shorting' the two switch wires. Of course that requires
you to remove the top of the meter.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Tuesday, January 15, 2013 4:30 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
I just noticed a gotcha, I need to set the key to calibrate before I may
issue those commands, I guess I need a key. I have printed off the photo and
measurements someone on this board did, now to get to a key shop...
M K
From: david.partridge@perdrix.co.uk
To: volt-nuts@febo.com
Date: Sun, 13 Jan 2013 17:14:07 +0000
Subject: Re: [volt-nuts] vSolartron7081 rom reading
If all you want is to save the calibration factors, just issue
CALIBRATE,DUMP command
To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH -
barring component failure that should be good for another ten years or so.
You can also display the NVM contents with the MONITOR command, but NOT
re-write them.
The only way that I am aware of that will write the caibration constants
if they have been lost is to run a full calibration.
Again all this in chapter 8.
I don't know which ROM programmers that can read or write the ER3400 (just
ignorance - there may be loads).
Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]
On Behalf Of m k
Sent: 13 January 2013 16:57
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading
Hi David,
well it is currently working and close to calibrated as far as I can tell.
But I want to get those calibration factors saved before it dies on me. Last
calibrated 2008, and well down on drift by now.
I get a diurnal 0.14 microvolt drift with a shorting plug that stays on
when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing
else to confirm anything else as I have only just got it chatting out
through the rs232.
I will have to buy something to read them with.
M K
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.