Bruce,
At 3:54 AM +0000 12/19/08, time-nuts-request@febo.com wrote:
Message: 6
Date: Fri, 19 Dec 2008 16:54:12 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] Sub Pico Second Phase logger
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Joe
Joe Gwinn wrote:
This is from home. I'll not be at work until next year.
At 11:48 PM +0000 12/18/08, time-nuts-request@febo.com wrote:
Message: 4
Date: Fri, 19 Dec 2008 12:17:33 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] Sub Pico Second Phase logger
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Joe
[snip]
Sources:
Wien bridge or equivalent (eg state variable oscillator with soft
clamping) relatively low distortion oscillator.
Buffered low pass filtered output of binary divider driven by a
crystal oscillator
RC oscillator sounds far simpler and more flexible.
A Wien bridge using a lamp is perhaps the simplest.
I'll create a circuit schematics for this using an OPA2134 (dual lowish
noise JFET opamp).
One opamp for the oscillator one to drive the sound card (attenuation of
the oscillator output will be required for some sound cards and it is
desirable to have a low output impedance driver).
Jim Williams of Linear Technology had a very good low-distortion AGC
controlled Wein Bridge oscillator. If I recall, he used a photo-FET
or the like as the servoed resistor in the bridge. There may be an
application note on the LT website, but I saw it in a chapter of a
book on analog circuitry, the chapter author being Jim W.
I have the application note.
LT App Note 43 is what JW mentions. Figure 47 is the key. (He also
mentions App Note 5.)
These cites are from the refs of "Max Wien, Mr. Hewlett, and a Rainy
Sunday Afternoon", Jim Williams, Chapter 7 in "Analog Circuit Design
-- Art, Science, and Personalities, edited by Jim Williams,
Butterworth-Heinemann 1991. The relevant section of App Note 43
seems to be shared with this chapter.
It wasn't a photo-FET he used, it was a Vactec model VTL5C10
optically (LED) driven CdS photocell.
Most of it is here (what's missing is some history and the refs):
http://books.google.com/books?id=Il4xxTTyhbEC&pg=PA43&lpg=PA43&dq=Max+Wien,+Mr.+Hewlett,+and+a+Rainy+Sunday+Afternoon&source=bl&ots=jnHNItJrOF&sig=ZL_CB_yIdT0ZbsD7Ieu_CjVYJHk&hl=en&sa=X&oi=book_result&resnum=1&ct=result#PPA53,M1.
For an AP192 the directly (without sample rate interpolation) available
output sample rates are:
192, 96, 64, 48, 32, 8 KSPS.
OK. I would start with 8 ksps. We will end up decimating below that
anyway, except for 1 KHz test signals.
Joe
I'll breadboard an oscillator over the holidays and take some samples.
A rainy-day activity for sure. In the Boston area, the issue is
snow - a noreaster is just now starting, with a foot of snow expected.
Joe
Joe
Joe Gwinn wrote:
Bruce,
At 3:54 AM +0000 12/19/08, time-nuts-request@febo.com wrote:
Message: 6
Date: Fri, 19 Dec 2008 16:54:12 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] Sub Pico Second Phase logger
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Joe
Joe Gwinn wrote:
This is from home. I'll not be at work until next year.
At 11:48 PM +0000 12/18/08, time-nuts-request@febo.com wrote:
Message: 4
Date: Fri, 19 Dec 2008 12:17:33 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] Sub Pico Second Phase logger
To: Discussion of precise time and frequency measurement
<time-nuts@febo.com>
Joe
[snip]
Sources:
Wien bridge or equivalent (eg state variable oscillator with soft
clamping) relatively low distortion oscillator.
Buffered low pass filtered output of binary divider driven by a
crystal oscillator
RC oscillator sounds far simpler and more flexible.
A Wien bridge using a lamp is perhaps the simplest.
I'll create a circuit schematics for this using an OPA2134 (dual lowish
noise JFET opamp).
One opamp for the oscillator one to drive the sound card (attenuation of
the oscillator output will be required for some sound cards and it is
desirable to have a low output impedance driver).
Jim Williams of Linear Technology had a very good low-distortion AGC
controlled Wein Bridge oscillator. If I recall, he used a photo-FET
or the like as the servoed resistor in the bridge. There may be an
application note on the LT website, but I saw it in a chapter of a
book on analog circuitry, the chapter author being Jim W.
I have the application note.
LT App Note 43 is what JW mentions. Figure 47 is the key. (He also
mentions App Note 5.)
These cites are from the refs of "Max Wien, Mr. Hewlett, and a Rainy
Sunday Afternoon", Jim Williams, Chapter 7 in "Analog Circuit Design
-- Art, Science, and Personalities, edited by Jim Williams,
Butterworth-Heinemann 1991. The relevant section of App Note 43
seems to be shared with this chapter.
It wasn't a photo-FET he used, it was a Vactec model VTL5C10
optically (LED) driven CdS photocell.
Note for those (in Australia, NZ and Europe) who may want to replicate
that oscillator:
The VTL5C10 device is obtainable locally from RS Components.
Most of it is here (what's missing is some history and the refs):
http://books.google.com/books?id=Il4xxTTyhbEC&pg=PA43&lpg=PA43&dq=Max+Wien,+Mr.+Hewlett,+and+a+Rainy+Sunday+Afternoon&source=bl&ots=jnHNItJrOF&sig=ZL_CB_yIdT0ZbsD7Ieu_CjVYJHk&hl=en&sa=X&oi=book_result&resnum=1&ct=result#PPA53,M1.
For an AP192 the directly (without sample rate interpolation) available
output sample rates are:
192, 96, 64, 48, 32, 8 KSPS.
OK. I would start with 8 ksps. We will end up decimating below that
anyway, except for 1 KHz test signals.
Joe
I'll breadboard an oscillator over the holidays and take some samples.
A rainy-day activity for sure. In the Boston area, the issue is
snow - a noreaster is just now starting, with a foot of snow expected.
Joe
AFAIK it has never snowed here.
It does snow on the central NI Plateau but not usually in December.
I'll start with a diode limited (soft clamps - resistance in series with
diodes) oscillator and move to a lower distortion one if necessary.
Bruce
Joe,
I have been following this thread from the start & learned a lot from it.
Thanks to you and Bruce for the insights.
I thought I would take a look at a simple, passive approach, just to
see what happens. I also limited my attention to stable 5 or 10 MHz
sources & observation windows of <100s. I used a Mini-circuits
SYPD-1 PD with a PSCQ-2-10.5 90 deg hybrid plus an ATM
P1506 phase shifter (delta=0.32ns). There are lots of other items
which will work just as well, this stuff happens to be on hand.
The SYPD-1 also needs a diplexer to terminate the high order
products & low pass filter the desired output; this is a cobble
together item from the parts box. I used a PC logging program
with my HP 3478A to take DC readings from the diplexer output.
Thus armed, I first checked the DVM logging with no RF in to see
if it was stable enough to yield a useable noise floor. This first result
was good enough. Reading 720 points over 100 seconds; the data
was limited to -1, 0 or +1 uV.
Then I applied 10MHz @ to the 90 deg hybrid & connected one
hybrid output to the SYPD-1 LO & the other hybrid output to the
ATM P1506. The ATM output then drives the SYPD-1 RF port.
I measured both SYPD-1 inputs to be +6.4dBm. Now I was able
to set the ATM phase to zero the SYPD-1 output & verify it was
stable over the 100s time I was interested in. The 3478A readings
ranged from -1 to +2uV during this time with STDDEV of 0.6uV
-again, good enough for me.
Finally, I used a HP 105B to get a quiet 5MHz source & a 1MHz
ext. reference for my HP3336C sythesizer. All this to give me 2
"nearly" identical. phase locked sources to measure the calibration
factor of the SYPD-1/Diplexer output. The 3336C was set to
5,000,000.005Hz & +6.4dBm driving the SYPD-1 LO port. The
105B 5MHz output was attenuated to +6.4dBm driving the RF
port. The resulting 5mHz data was logged & analyzed at the zero
crossings (max gain). The result is a phase sensitivity of 1.22uRad/uV.
The + & - zero crossings were identical and the 1.22uRad/uV factor
is linear (+/- 2 %) up to +/- 20 % of FS. FS = 1.023V
If my math is about right, @10MHz, 1uRad = 16fs; so a noise floor
of +/- 2uV (+/-2.5uRad) equals +/-40fs. I expect that, if 2 sources
are near zero phase, then it should only require <10s to estimate
the actual phase error within 1ps, or less. Anyone agree?
I await & hope for constructive comments and/or corrections.
Pete Rawson
Pete
Pete wrote:
Joe,
I have been following this thread from the start & learned a lot from it.
Thanks to you and Bruce for the insights.
I thought I would take a look at a simple, passive approach, just to
see what happens. I also limited my attention to stable 5 or 10 MHz
sources & observation windows of <100s. I used a Mini-circuits
SYPD-1 PD with a PSCQ-2-10.5 90 deg hybrid plus an ATM
P1506 phase shifter (delta=0.32ns). There are lots of other items
which will work just as well, this stuff happens to be on hand.
The SYPD-1 also needs a diplexer to terminate the high order
products & low pass filter the desired output; this is a cobble
together item from the parts box. I used a PC logging program
with my HP 3478A to take DC readings from the diplexer output.
The use of a diplexer perpetuates a common fallacy.
You get significantly lower noise with a capacitive IF port termination.
This is true even in the microwave region when reflecting the sum
frequency back into the mixer reduces the noise.
At low frequencies this has little effect on the Rf and LO port VSWR.
The capacitive termination limits the IF port bandwidth to a few tens of
kHz and a 50 ohm termination is more useful if an IF bandwidth of
several MHz is desired.
Thus armed, I first checked the DVM logging with no RF in to see
if it was stable enough to yield a useable noise floor. This first result
was good enough. Reading 720 points over 100 seconds; the data
was limited to -1, 0 or +1 uV.
Then I applied 10MHz @ to the 90 deg hybrid & connected one
hybrid output to the SYPD-1 LO & the other hybrid output to the
ATM P1506. The ATM output then drives the SYPD-1 RF port.
I measured both SYPD-1 inputs to be +6.4dBm. Now I was able
to set the ATM phase to zero the SYPD-1 output & verify it was
stable over the 100s time I was interested in. The 3478A readings
ranged from -1 to +2uV during this time with STDDEV of 0.6uV
-again, good enough for me.
Can you produce ADEV and MDEV plots for tau up to 1000sec or more?
This would then give us some idea of the noise floor of your setup.
SDEV by itself can be very misleading as for some noise processes it
diverges.
The RPD and MPD series are better, in that they allow low frequency
isolation of the RF, LO and IF port grounds.
This can be critical for low offset frequency measurements.
Finally, I used a HP 105B to get a quiet 5MHz source & a 1MHz
ext. reference for my HP3336C sythesizer. All this to give me 2
"nearly" identical. phase locked sources to measure the calibration
factor of the SYPD-1/Diplexer output. The 3336C was set to
5,000,000.005Hz & +6.4dBm driving the SYPD-1 LO port. The
105B 5MHz output was attenuated to +6.4dBm driving the RF
port. The resulting 5mHz data was logged & analyzed at the zero
crossings (max gain). The result is a phase sensitivity of 1.22uRad/uV.
The + & - zero crossings were identical and the 1.22uRad/uV factor
is linear (+/- 2 %) up to +/- 20 % of FS. FS = 1.023V
If my math is about right, @10MHz, 1uRad = 16fs; so a noise floor
of +/- 2uV (+/-2.5uRad) equals +/-40fs. I expect that, if 2 sources
are near zero phase, then it should only require <10s to estimate
the actual phase error within 1ps, or less. Anyone agree?
I await & hope for constructive comments and/or corrections.
Pete Rawson
At 10MHz 1 radian = 16ns
1 millirad = 16 ps.
1urad = 16fs.
It would be more useful if you logged the DVM readings for an hour or
more so that ADEV, MDEV plots can be produced.
Eyeballing the data doesnt always deliver the full story.
If one compares a couple of oscillators one has to be very careful to
ensure that inadvertent injection locking isn't occuring.
This can happen with almost imperceptible coupling if the oscillators
are very close in frequency.
You need to use very high reverse isolation ( ideally 150dB or more)
buffer amplifiers to ensure this doesnt occur.
The first test you made only indicates the system noise floor.
When the mixer/phase detector is driven by 2 different oscillators the
noise level will be much higher that the noise level of the mixer system
itself.
In fact practically all OCXOs will/should be much noisier.
If the noise is too low this is an indication of injection locking.
One thing to watch out for is the measurement system noise bandwidth.
You need to know this as ADEV, MDEV etc depend on the actual noise
bandwidth.
When you use a DVM it usually has built in low pass filtering as well as
its inherent sinc response due to the integration process.
It can be useful to use external passive components to accurately define
the noise response.
Bruce
Pete
Another very useful thing to do is to measure the output noise spectrum
at the of mixer port both with capacitive termination and with a diplexer.
You can do this with your PC sound card system (even the motherboard
one) and a low noise preamp.
If you want a simple low noise preamp circuit I can supply several.
The gain required depends on the sound card depends on the sound cards
internal noise level.
One way is to drive the RF and LO ports with quadrature phase signals
and measure the IF spectrum.
Another is to just drive the LO port and look at the IF port spectrum.
Try both with the Rf port open circuit and terminated in 50 ohms.
Once you've obtained spectra posting them would be useful and informative.
Bruce
Bruce Griffiths wrote:
Pete
Pete wrote:
Joe,
I have been following this thread from the start & learned a lot from it.
Thanks to you and Bruce for the insights.
I thought I would take a look at a simple, passive approach, just to
see what happens. I also limited my attention to stable 5 or 10 MHz
sources & observation windows of <100s. I used a Mini-circuits
SYPD-1 PD with a PSCQ-2-10.5 90 deg hybrid plus an ATM
P1506 phase shifter (delta=0.32ns). There are lots of other items
which will work just as well, this stuff happens to be on hand.
The SYPD-1 also needs a diplexer to terminate the high order
products & low pass filter the desired output; this is a cobble
together item from the parts box. I used a PC logging program
with my HP 3478A to take DC readings from the diplexer output.
The use of a diplexer perpetuates a common fallacy.
You get significantly lower noise with a capacitive IF port termination.
This is true even in the microwave region when reflecting the sum
frequency back into the mixer reduces the noise.
At low frequencies this has little effect on the Rf and LO port VSWR.
The capacitive termination limits the IF port bandwidth to a few tens of
kHz and a 50 ohm termination is more useful if an IF bandwidth of
several MHz is desired.
Thus armed, I first checked the DVM logging with no RF in to see
if it was stable enough to yield a useable noise floor. This first result
was good enough. Reading 720 points over 100 seconds; the data
was limited to -1, 0 or +1 uV.
Then I applied 10MHz @ to the 90 deg hybrid & connected one
hybrid output to the SYPD-1 LO & the other hybrid output to the
ATM P1506. The ATM output then drives the SYPD-1 RF port.
I measured both SYPD-1 inputs to be +6.4dBm. Now I was able
to set the ATM phase to zero the SYPD-1 output & verify it was
stable over the 100s time I was interested in. The 3478A readings
ranged from -1 to +2uV during this time with STDDEV of 0.6uV
-again, good enough for me.
Can you produce ADEV and MDEV plots for tau up to 1000sec or more?
This would then give us some idea of the noise floor of your setup.
SDEV by itself can be very misleading as for some noise processes it
diverges.
The RPD and MPD series are better, in that they allow low frequency
isolation of the RF, LO and IF port grounds.
This can be critical for low offset frequency measurements.
Finally, I used a HP 105B to get a quiet 5MHz source & a 1MHz
ext. reference for my HP3336C sythesizer. All this to give me 2
"nearly" identical. phase locked sources to measure the calibration
factor of the SYPD-1/Diplexer output. The 3336C was set to
5,000,000.005Hz & +6.4dBm driving the SYPD-1 LO port. The
105B 5MHz output was attenuated to +6.4dBm driving the RF
port. The resulting 5mHz data was logged & analyzed at the zero
crossings (max gain). The result is a phase sensitivity of 1.22uRad/uV.
The + & - zero crossings were identical and the 1.22uRad/uV factor
is linear (+/- 2 %) up to +/- 20 % of FS. FS = 1.023V
If my math is about right, @10MHz, 1uRad = 16fs; so a noise floor
of +/- 2uV (+/-2.5uRad) equals +/-40fs. I expect that, if 2 sources
are near zero phase, then it should only require <10s to estimate
the actual phase error within 1ps, or less. Anyone agree?
I await & hope for constructive comments and/or corrections.
Pete Rawson
At 10MHz 1 radian = 16ns
1 millirad = 16 ps.
1urad = 16fs.
It would be more useful if you logged the DVM readings for an hour or
more so that ADEV, MDEV plots can be produced.
Eyeballing the data doesnt always deliver the full story.
If one compares a couple of oscillators one has to be very careful to
ensure that inadvertent injection locking isn't occuring.
This can happen with almost imperceptible coupling if the oscillators
are very close in frequency.
You need to use very high reverse isolation ( ideally 150dB or more)
buffer amplifiers to ensure this doesnt occur.
The first test you made only indicates the system noise floor.
When the mixer/phase detector is driven by 2 different oscillators the
noise level will be much higher that the noise level of the mixer system
itself.
In fact practically all OCXOs will/should be much noisier.
If the noise is too low this is an indication of injection locking.
One thing to watch out for is the measurement system noise bandwidth.
You need to know this as ADEV, MDEV etc depend on the actual noise
bandwidth.
When you use a DVM it usually has built in low pass filtering as well as
its inherent sinc response due to the integration process.
It can be useful to use external passive components to accurately define
the noise response.
Bruce