usrp-users@lists.ettus.com

Discussion and technical support related to USRP, UHD, RFNoC

View all threads

N310: Channel pairs are drifting in phase whith external LO setup

JB
Janos Buttgereit
Fri, Mar 22, 2019 9:31 AM

Hi,

after having struggled a lot with network config issues I finally got out N310 up and running. We chosed it for an application that needs perfectly constant phases between all four channels to build up DOA estimation applications.

I’ve set up an N210 as a signal source, outputting a sine round about 1.5GHz feeding a 4-way splitter that than feeds all four input channels of the N310. A Rhode&Schwarz signal generator outputting a 3GHz sine wave feeds a two way splitter that is connected to both external rx LO inputs.

From this setup I expect all four channels to maybe have a static phase offset that does not change over time. However, it can be noticed that the pairs of channels 0 and 1 and the pairs of channels 2 and 3 show those behavior but both pairs drift against each other over time. Any idea of what could be wrong with this setup?

I attached the GNURadio flow graph for this setup.

Thanks in Advance!

Hi, after having struggled a lot with network config issues I finally got out N310 up and running. We chosed it for an application that needs perfectly constant phases between all four channels to build up DOA estimation applications. I’ve set up an N210 as a signal source, outputting a sine round about 1.5GHz feeding a 4-way splitter that than feeds all four input channels of the N310. A Rhode&Schwarz signal generator outputting a 3GHz sine wave feeds a two way splitter that is connected to both external rx LO inputs. From this setup I expect all four channels to maybe have a static phase offset that does not change over time. However, it can be noticed that the pairs of channels 0 and 1 and the pairs of channels 2 and 3 show those behavior but both pairs drift against each other over time. Any idea of what could be wrong with this setup? I attached the GNURadio flow graph for this setup. Thanks in Advance!
MD
Marcus D. Leech
Fri, Mar 22, 2019 1:17 PM

On 03/22/2019 05:31 AM, Janos Buttgereit via USRP-users wrote:

Hi,

after having struggled a lot with network config issues I finally got out N310 up and running. We chosed it for an application that needs perfectly constant phases between all four channels to build up DOA estimation applications.

I’ve set up an N210 as a signal source, outputting a sine round about 1.5GHz feeding a 4-way splitter that than feeds all four input channels of the N310. A Rhode&Schwarz signal generator outputting a 3GHz sine wave feeds a two way splitter that is connected to both external rx LO inputs.

From this setup I expect all four channels to maybe have a static phase offset that does not change over time. However, it can be noticed that the pairs of channels 0 and 1 and the pairs of channels 2 and 3 show those behavior but both pairs drift against each other over time. Any idea of what could be wrong with this setup?

I attached the GNURadio flow graph for this setup.

Thanks in Advance!

Change the "Sync" option to "Unknown PPS".  Why this is required is a
bit of a mystery, but another user experienced the same issue.

The N310 is conceptually two radios in a single box, and each radio
within the FPGA has its own "timekeeper".  For some reason, unless those
"timekeepers" are started synchronized with each other, sample
timestamps drift relative to one another.  This should not be the case, but
it appears to be.

On 03/22/2019 05:31 AM, Janos Buttgereit via USRP-users wrote: > Hi, > > after having struggled a lot with network config issues I finally got out N310 up and running. We chosed it for an application that needs perfectly constant phases between all four channels to build up DOA estimation applications. > > I’ve set up an N210 as a signal source, outputting a sine round about 1.5GHz feeding a 4-way splitter that than feeds all four input channels of the N310. A Rhode&Schwarz signal generator outputting a 3GHz sine wave feeds a two way splitter that is connected to both external rx LO inputs. > > From this setup I expect all four channels to maybe have a static phase offset that does not change over time. However, it can be noticed that the pairs of channels 0 and 1 and the pairs of channels 2 and 3 show those behavior but both pairs drift against each other over time. Any idea of what could be wrong with this setup? > > I attached the GNURadio flow graph for this setup. > > Thanks in Advance! > > Change the "Sync" option to "Unknown PPS". Why this is required is a bit of a mystery, but another user experienced the same issue. The N310 is conceptually two radios in a single box, and each radio within the FPGA has its own "timekeeper". For some reason, unless those "timekeepers" are started synchronized with each other, sample timestamps drift relative to one another. This should not be the case, but it appears to be.
JB
Janos Buttgereit
Fri, Mar 22, 2019 2:40 PM

Thank you for the quick reply, that did the trick, at least half way.

Now I’m struggling with the random 180° phase offset between the front end pairs, obviously generated by the AD9371 chip according to my research. Is there any best-practice in detecting this phase shift that does not involve an additional reference signal source?

Am 22.03.2019 um 14:17 schrieb Marcus D. Leech via USRP-users usrp-users@lists.ettus.com:

On 03/22/2019 05:31 AM, Janos Buttgereit via USRP-users wrote:

Hi,

after having struggled a lot with network config issues I finally got out N310 up and running. We chosed it for an application that needs perfectly constant phases between all four channels to build up DOA estimation applications.

I’ve set up an N210 as a signal source, outputting a sine round about 1.5GHz feeding a 4-way splitter that than feeds all four input channels of the N310. A Rhode&Schwarz signal generator outputting a 3GHz sine wave feeds a two way splitter that is connected to both external rx LO inputs.

From this setup I expect all four channels to maybe have a static phase offset that does not change over time. However, it can be noticed that the pairs of channels 0 and 1 and the pairs of channels 2 and 3 show those behavior but both pairs drift against each other over time. Any idea of what could be wrong with this setup?

I attached the GNURadio flow graph for this setup.

Thanks in Advance!

Change the "Sync" option to "Unknown PPS".  Why this is required is a bit of a mystery, but another user experienced the same issue.

The N310 is conceptually two radios in a single box, and each radio within the FPGA has its own "timekeeper".  For some reason, unless those
"timekeepers" are started synchronized with each other, sample timestamps drift relative to one another.  This should not be the case, but
it appears to be.


USRP-users mailing list
USRP-users@lists.ettus.com mailto:USRP-users@lists.ettus.com
http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com

Thank you for the quick reply, that did the trick, at least half way. Now I’m struggling with the random 180° phase offset between the front end pairs, obviously generated by the AD9371 chip according to my research. Is there any best-practice in detecting this phase shift that does not involve an additional reference signal source? > Am 22.03.2019 um 14:17 schrieb Marcus D. Leech via USRP-users <usrp-users@lists.ettus.com>: > > On 03/22/2019 05:31 AM, Janos Buttgereit via USRP-users wrote: >> Hi, >> >> after having struggled a lot with network config issues I finally got out N310 up and running. We chosed it for an application that needs perfectly constant phases between all four channels to build up DOA estimation applications. >> >> I’ve set up an N210 as a signal source, outputting a sine round about 1.5GHz feeding a 4-way splitter that than feeds all four input channels of the N310. A Rhode&Schwarz signal generator outputting a 3GHz sine wave feeds a two way splitter that is connected to both external rx LO inputs. >> >> From this setup I expect all four channels to maybe have a static phase offset that does not change over time. However, it can be noticed that the pairs of channels 0 and 1 and the pairs of channels 2 and 3 show those behavior but both pairs drift against each other over time. Any idea of what could be wrong with this setup? >> >> I attached the GNURadio flow graph for this setup. >> >> Thanks in Advance! >> > Change the "Sync" option to "Unknown PPS". Why this is required is a bit of a mystery, but another user experienced the same issue. > > The N310 is conceptually two radios in a single box, and each radio within the FPGA has its own "timekeeper". For some reason, unless those > "timekeepers" are started synchronized with each other, sample timestamps drift relative to one another. This should not be the case, but > it appears to be. > > > > > _______________________________________________ > USRP-users mailing list > USRP-users@lists.ettus.com <mailto:USRP-users@lists.ettus.com> > http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com <http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com>
MD
Marcus D. Leech
Fri, Mar 22, 2019 2:46 PM

On 03/22/2019 10:40 AM, Janos Buttgereit via USRP-users wrote:

Thank you for the quick reply, that did the trick, at least half way.

Now I’m struggling with the random 180° phase offset between the front
end pairs, obviously generated by the AD9371 chip according to my
research. Is there any best-practice in detecting this phase shift
that does not involve an additional reference signal source?

Correction by analysis of the RX stream is the way it is done.

In the specific case of 180deg, one "guess" at cross-correlating the
channels will produce a louder cross product than the other--you only
have two guesses that you need to make.

It is generally exceedingly difficult to make a collection of N PLL
synthesized LOs all have zero relative phase offset, unless those
synthesizers
have "special features" to allow phase-resynch.

So, a lot of applications simply have algorithms for dealing with
achieving phase alignment within the RX processing.

Am 22.03.2019 um 14:17 schrieb Marcus D. Leech via USRP-users
<usrp-users@lists.ettus.com mailto:usrp-users@lists.ettus.com>:

On 03/22/2019 05:31 AM, Janos Buttgereit via USRP-users wrote:

Hi,

after having struggled a lot with network config issues I finally
got out N310 up and running. We chosed it for an application that
needs perfectly constant phases between all four channels to build
up DOA estimation applications.

I’ve set up an N210 as a signal source, outputting a sine round
about 1.5GHz feeding a 4-way splitter that than feeds all four input
channels of the N310. A Rhode&Schwarz signal generator outputting a
3GHz sine wave feeds a two way splitter that is connected to both
external rx LO inputs.

From this setup I expect all four channels to maybe have a static
phase offset that does not change over time. However, it can be
noticed that the pairs of channels 0 and 1 and the pairs of channels
2 and 3 show those behavior but both pairs drift against each other
over time. Any idea of what could be wrong with this setup?

I attached the GNURadio flow graph for this setup.

Thanks in Advance!

Change the "Sync" option to "Unknown PPS".  Why this is required is
a bit of a mystery, but another user experienced the same issue.

The N310 is conceptually two radios in a single box, and each radio
within the FPGA has its own "timekeeper".  For some reason, unless those
"timekeepers" are started synchronized with each other, sample
timestamps drift relative to one another.  This should not be the
case, but
it appears to be.


USRP-users mailing list
USRP-users@lists.ettus.com mailto:USRP-users@lists.ettus.com
http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com

On 03/22/2019 10:40 AM, Janos Buttgereit via USRP-users wrote: > Thank you for the quick reply, that did the trick, at least half way. > > Now I’m struggling with the random 180° phase offset between the front > end pairs, obviously generated by the AD9371 chip according to my > research. Is there any best-practice in detecting this phase shift > that does not involve an additional reference signal source? Correction by analysis of the RX stream is the way it is done. In the specific case of 180deg, one "guess" at cross-correlating the channels will produce a louder cross product than the other--you only have two guesses that you need to make. It is generally exceedingly difficult to make a collection of N PLL synthesized LOs all have zero relative phase offset, unless those synthesizers have "special features" to allow phase-resynch. So, a lot of applications simply have algorithms for dealing with achieving phase alignment within the RX processing. > >> Am 22.03.2019 um 14:17 schrieb Marcus D. Leech via USRP-users >> <usrp-users@lists.ettus.com <mailto:usrp-users@lists.ettus.com>>: >> >> On 03/22/2019 05:31 AM, Janos Buttgereit via USRP-users wrote: >>> Hi, >>> >>> after having struggled a lot with network config issues I finally >>> got out N310 up and running. We chosed it for an application that >>> needs perfectly constant phases between all four channels to build >>> up DOA estimation applications. >>> >>> I’ve set up an N210 as a signal source, outputting a sine round >>> about 1.5GHz feeding a 4-way splitter that than feeds all four input >>> channels of the N310. A Rhode&Schwarz signal generator outputting a >>> 3GHz sine wave feeds a two way splitter that is connected to both >>> external rx LO inputs. >>> >>> From this setup I expect all four channels to maybe have a static >>> phase offset that does not change over time. However, it can be >>> noticed that the pairs of channels 0 and 1 and the pairs of channels >>> 2 and 3 show those behavior but both pairs drift against each other >>> over time. Any idea of what could be wrong with this setup? >>> >>> I attached the GNURadio flow graph for this setup. >>> >>> Thanks in Advance! >>> >> Change the "Sync" option to "Unknown PPS". Why this is required is >> a bit of a mystery, but another user experienced the same issue. >> >> The N310 is conceptually two radios in a single box, and each radio >> within the FPGA has its own "timekeeper". For some reason, unless those >> "timekeepers" are started synchronized with each other, sample >> timestamps drift relative to one another. This should not be the >> case, but >> it appears to be. >> >> >> >> >> _______________________________________________ >> USRP-users mailing list >> USRP-users@lists.ettus.com <mailto:USRP-users@lists.ettus.com> >> http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com > > > > _______________________________________________ > USRP-users mailing list > USRP-users@lists.ettus.com > http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com