volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

HP-3458A questions

FS
Frank Stellmach
Tue, Jun 2, 2015 9:50 PM

Hi Randy,

in advance, this change only improves the stability of the instrument,
when the 3458A is powered.
The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an
extended period of time.

(Reference: Spreadbury: "Ultra Zener..")

If you change the oven temperature, you will loose the 10V calibration,
as the LTZ1000A will change its reference voltage.

Although my instrument is strictly switched off, when not doing
measurements, I reduced the oven temp. by connecting a 100k precision
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k
in total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP,
it's component designator is X411.
I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

I would recommend my solution above the one of lymex, although his one
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t)
of the 15k BMF.

With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the
fan has ALWAYS to be kept clean, giving a maximum temperature rise of
15°C, and the instrument may not be put in a rack (which would give
additional 10..20°C temperature rise). Under these conditions, (35°C
ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C
regulation margin), the oven will always be stable.

These 75k / 55°C would be possible, only if you limit the ambient
temperature further, to <= 25°C, and precisely monitor the 3458A inner
temperature (TEMP?).

Although my instrument is kept at 20..23°C throughout the year, i
wouldn't feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A
version, as the self heating effect may vanish!

Frank

Hi Randy, in advance, this change only improves the stability of the instrument, when the 3458A is powered. The LTZ1000 will not age (nearly), when the instrument is switched off. Therefore, it makes sense mainly, if your instrument is running for an extended period of time. (Reference: Spreadbury: "Ultra Zener..") If you change the oven temperature, you will loose the 10V calibration, as the LTZ1000A will change its reference voltage. Although my instrument is strictly switched off, when not doing measurements, I reduced the oven temp. by connecting a 100k precision wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k in total, yielding about 65°C like in the LT datasheet. An alternative placement is already foreseen on the PCB; in the CLIP, it's component designator is X411. I'll try to link in a photo of my change. lymex on the bbs35.hot forum obviously was more agressive, and paralled 75k of Vitrohm Thin Film, giving 12.5k for about 55°C. I would recommend my solution above the one of lymex, although his one gives theoretically more stability. Also, the lower stability of TF reistors may affect the stability (T, t) of the 15k BMF. With 100k/65°C, the oven runs about 30°C lower. Therefore, the ambient temperature has to be limited to max. 35°C, the fan has ALWAYS to be kept clean, giving a maximum temperature rise of 15°C, and the instrument may not be put in a rack (which would give additional 10..20°C temperature rise). Under these conditions, (35°C ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C regulation margin), the oven will always be stable. These 75k / 55°C would be possible, only if you limit the ambient temperature further, to <= 25°C, and precisely monitor the 3458A inner temperature (TEMP?). Although my instrument is kept at 20..23°C throughout the year, i wouldn't feel comfortable with 55°C. The only possibility would be the exchange of the reference to the non-A version, as the self heating effect may vanish! Frank
JL
J. L. Trantham
Tue, Jun 2, 2015 11:00 PM

Frank,

Thanks for the picture of your board.

I have been interested in the 'plastic cap' on the LTZ1000A and you provided me with information about this in the past.

Do you have a picture of the bottom of the board?  I would love to see how the bottom half of the 'cap' looks.

Thanks.

Joe

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Frank Stellmach
Sent: Tuesday, June 02, 2015 4:50 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP-3458A questions

Hi Randy,

in advance, this change only improves the stability of the instrument, when the 3458A is powered.
The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an extended period of time.

(Reference: Spreadbury: "Ultra Zener..")

If you change the oven temperature, you will loose the 10V calibration, as the LTZ1000A will change its reference voltage.

Although my instrument is strictly switched off, when not doing
measurements, I reduced the oven temp. by connecting a 100k precision
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k
in total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP,
it's component designator is X411.
I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

I would recommend my solution above the one of lymex, although his one
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t)
of the 15k BMF.

With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the
fan has ALWAYS to be kept clean, giving a maximum temperature rise of
15°C, and the instrument may not be put in a rack (which would give
additional 10..20°C temperature rise). Under these conditions, (35°C
ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C
regulation margin), the oven will always be stable.

These 75k / 55°C would be possible, only if you limit the ambient
temperature further, to <= 25°C, and precisely monitor the 3458A inner
temperature (TEMP?).

Although my instrument is kept at 20..23°C throughout the year, i
wouldn't feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A
version, as the self heating effect may vanish!

Frank

Frank, Thanks for the picture of your board. I have been interested in the 'plastic cap' on the LTZ1000A and you provided me with information about this in the past. Do you have a picture of the bottom of the board? I would love to see how the bottom half of the 'cap' looks. Thanks. Joe -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Frank Stellmach Sent: Tuesday, June 02, 2015 4:50 PM To: volt-nuts@febo.com Subject: [volt-nuts] HP-3458A questions Hi Randy, in advance, this change only improves the stability of the instrument, when the 3458A is powered. The LTZ1000 will not age (nearly), when the instrument is switched off. Therefore, it makes sense mainly, if your instrument is running for an extended period of time. (Reference: Spreadbury: "Ultra Zener..") If you change the oven temperature, you will loose the 10V calibration, as the LTZ1000A will change its reference voltage. Although my instrument is strictly switched off, when not doing measurements, I reduced the oven temp. by connecting a 100k precision wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k in total, yielding about 65°C like in the LT datasheet. An alternative placement is already foreseen on the PCB; in the CLIP, it's component designator is X411. I'll try to link in a photo of my change. lymex on the bbs35.hot forum obviously was more agressive, and paralled 75k of Vitrohm Thin Film, giving 12.5k for about 55°C. I would recommend my solution above the one of lymex, although his one gives theoretically more stability. Also, the lower stability of TF reistors may affect the stability (T, t) of the 15k BMF. With 100k/65°C, the oven runs about 30°C lower. Therefore, the ambient temperature has to be limited to max. 35°C, the fan has ALWAYS to be kept clean, giving a maximum temperature rise of 15°C, and the instrument may not be put in a rack (which would give additional 10..20°C temperature rise). Under these conditions, (35°C ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C regulation margin), the oven will always be stable. These 75k / 55°C would be possible, only if you limit the ambient temperature further, to <= 25°C, and precisely monitor the 3458A inner temperature (TEMP?). Although my instrument is kept at 20..23°C throughout the year, i wouldn't feel comfortable with 55°C. The only possibility would be the exchange of the reference to the non-A version, as the self heating effect may vanish! Frank
AJ
Andreas Jahn
Wed, Jun 3, 2015 3:51 AM

Hello,

A bottom picture of the board is currently in a auction (so no need to
deinstall the board from a life system):
http://www.ebay.com/itm/Agilent-03458-66509-DCV-Reference-PCA-for-3458-PART-/371306381959?pt=LH_DefaultDomain_0&hash=item567395c287

With best regards

Andreas

Am 03.06.2015 um 01:00 schrieb J. L. Trantham:

Frank,

Thanks for the picture of your board.

I have been interested in the 'plastic cap' on the LTZ1000A and you provided me with information about this in the past.

Do you have a picture of the bottom of the board?  I would love to see how the bottom half of the 'cap' looks.

Thanks.

Joe

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Frank Stellmach
Sent: Tuesday, June 02, 2015 4:50 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP-3458A questions

Hi Randy,

in advance, this change only improves the stability of the instrument, when the 3458A is powered.
The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an extended period of time.

(Reference: Spreadbury: "Ultra Zener..")

If you change the oven temperature, you will loose the 10V calibration, as the LTZ1000A will change its reference voltage.

Although my instrument is strictly switched off, when not doing
measurements, I reduced the oven temp. by connecting a 100k precision
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k
in total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP,
it's component designator is X411.
I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

I would recommend my solution above the one of lymex, although his one
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t)
of the 15k BMF.

With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the
fan has ALWAYS to be kept clean, giving a maximum temperature rise of
15°C, and the instrument may not be put in a rack (which would give
additional 10..20°C temperature rise). Under these conditions, (35°C
ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C
regulation margin), the oven will always be stable.

These 75k / 55°C would be possible, only if you limit the ambient
temperature further, to <= 25°C, and precisely monitor the 3458A inner
temperature (TEMP?).

Although my instrument is kept at 20..23°C throughout the year, i
wouldn't feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A
version, as the self heating effect may vanish!

Frank


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hello, A bottom picture of the board is currently in a auction (so no need to deinstall the board from a life system): http://www.ebay.com/itm/Agilent-03458-66509-DCV-Reference-PCA-for-3458-PART-/371306381959?pt=LH_DefaultDomain_0&hash=item567395c287 With best regards Andreas Am 03.06.2015 um 01:00 schrieb J. L. Trantham: > Frank, > > Thanks for the picture of your board. > > I have been interested in the 'plastic cap' on the LTZ1000A and you provided me with information about this in the past. > > Do you have a picture of the bottom of the board? I would love to see how the bottom half of the 'cap' looks. > > Thanks. > > Joe > > -----Original Message----- > From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Frank Stellmach > Sent: Tuesday, June 02, 2015 4:50 PM > To: volt-nuts@febo.com > Subject: [volt-nuts] HP-3458A questions > > Hi Randy, > > in advance, this change only improves the stability of the instrument, when the 3458A is powered. > The LTZ1000 will not age (nearly), when the instrument is switched off. > Therefore, it makes sense mainly, if your instrument is running for an extended period of time. > > (Reference: Spreadbury: "Ultra Zener..") > > If you change the oven temperature, you will loose the 10V calibration, as the LTZ1000A will change its reference voltage. > > > Although my instrument is strictly switched off, when not doing > measurements, I reduced the oven temp. by connecting a 100k precision > wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k > in total, yielding about 65°C like in the LT datasheet. > An alternative placement is already foreseen on the PCB; in the CLIP, > it's component designator is X411. > I'll try to link in a photo of my change. > > lymex on the bbs35.hot forum obviously was more agressive, and paralled > 75k of Vitrohm Thin Film, giving 12.5k for about 55°C. > > I would recommend my solution above the one of lymex, although his one > gives theoretically more stability. > Also, the lower stability of TF reistors may affect the stability (T, t) > of the 15k BMF. > > With 100k/65°C, the oven runs about 30°C lower. > Therefore, the ambient temperature has to be limited to max. 35°C, the > fan has ALWAYS to be kept clean, giving a maximum temperature rise of > 15°C, and the instrument may not be put in a rack (which would give > additional 10..20°C temperature rise). Under these conditions, (35°C > ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C > regulation margin), the oven will always be stable. > > These 75k / 55°C would be possible, only if you limit the ambient > temperature further, to <= 25°C, and precisely monitor the 3458A inner > temperature (TEMP?). > > Although my instrument is kept at 20..23°C throughout the year, i > wouldn't feel comfortable with 55°C. > The only possibility would be the exchange of the reference to the non-A > version, as the self heating effect may vanish! > > Frank > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
RE
Randy Evans
Thu, Jun 4, 2015 1:07 AM

Frank,

Thanks for the data.  The 100K seems acceptable since I have a Fluke 732A
to recalibrate the 3458A with after the change and the 3458A is not in a
rack.

Randy

On Tue, Jun 2, 2015 at 2:50 PM, Frank Stellmach frank.stellmach@freenet.de
wrote:

Hi Randy,

in advance, this change only improves the stability of the instrument,
when the 3458A is powered.
The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an
extended period of time.

(Reference: Spreadbury: "Ultra Zener..")

If you change the oven temperature, you will loose the 10V calibration, as
the LTZ1000A will change its reference voltage.

Although my instrument is strictly switched off, when not doing
measurements, I reduced the oven temp. by connecting a 100k precision
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k in
total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP, it's
component designator is X411.
I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

I would recommend my solution above the one of lymex, although his one
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t)
of the 15k BMF.

With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the fan
has ALWAYS to be kept clean, giving a maximum temperature rise of 15°C, and
the instrument may not be put in a rack (which would give additional
10..20°C temperature rise). Under these conditions, (35°C ambient + 15°C
3458A inner heating + 10°C self heating of LTZ1000A + 5°C regulation
margin), the oven will always be stable.

These 75k / 55°C would be possible, only if you limit the ambient
temperature further, to <= 25°C, and precisely monitor the 3458A inner
temperature (TEMP?).

Although my instrument is kept at 20..23°C throughout the year, i wouldn't
feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A
version, as the self heating effect may vanish!

Frank


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Frank, Thanks for the data. The 100K seems acceptable since I have a Fluke 732A to recalibrate the 3458A with after the change and the 3458A is not in a rack. Randy On Tue, Jun 2, 2015 at 2:50 PM, Frank Stellmach <frank.stellmach@freenet.de> wrote: > Hi Randy, > > in advance, this change only improves the stability of the instrument, > when the 3458A is powered. > The LTZ1000 will not age (nearly), when the instrument is switched off. > Therefore, it makes sense mainly, if your instrument is running for an > extended period of time. > > (Reference: Spreadbury: "Ultra Zener..") > > If you change the oven temperature, you will loose the 10V calibration, as > the LTZ1000A will change its reference voltage. > > > Although my instrument is strictly switched off, when not doing > measurements, I reduced the oven temp. by connecting a 100k precision > wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k in > total, yielding about 65°C like in the LT datasheet. > An alternative placement is already foreseen on the PCB; in the CLIP, it's > component designator is X411. > I'll try to link in a photo of my change. > > lymex on the bbs35.hot forum obviously was more agressive, and paralled > 75k of Vitrohm Thin Film, giving 12.5k for about 55°C. > > I would recommend my solution above the one of lymex, although his one > gives theoretically more stability. > Also, the lower stability of TF reistors may affect the stability (T, t) > of the 15k BMF. > > With 100k/65°C, the oven runs about 30°C lower. > Therefore, the ambient temperature has to be limited to max. 35°C, the fan > has ALWAYS to be kept clean, giving a maximum temperature rise of 15°C, and > the instrument may not be put in a rack (which would give additional > 10..20°C temperature rise). Under these conditions, (35°C ambient + 15°C > 3458A inner heating + 10°C self heating of LTZ1000A + 5°C regulation > margin), the oven will always be stable. > > These 75k / 55°C would be possible, only if you limit the ambient > temperature further, to <= 25°C, and precisely monitor the 3458A inner > temperature (TEMP?). > > Although my instrument is kept at 20..23°C throughout the year, i wouldn't > feel comfortable with 55°C. > The only possibility would be the exchange of the reference to the non-A > version, as the self heating effect may vanish! > > Frank > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >