DMTD phase shifter

CD
Corby Dawson
Fri, Jul 24, 2009 5:28 PM

Stephan,

I have a DMTD system that was built by NBS (now NIST) in the early 80s.

The coarse phase shifter is made up of ten lengths of miniature coax
cable bundled up in a shielded box with toggle switches on the front
panel to select the delay used.

The LSB is about 2us.

It also has a fine adjust that consists of a couple varactors in a phase
delay circuit.

I have found that with a good quality offset L.O. I can get the precision
I need (<1X10-13th at a 1 second interval) without using the fine adjust
circuit.

As far as the small time interval value I usually start my measurements
at around 2 to 4 us delay with the DUT set for a slowly increasing phase
shift,

The accuracy required depends on what your measuring but with a 10Mhz
input and a 1hz beat note +-1us is equal to +-1X10-13th so most counters
will work just fine. Just be advised that the us and below digits will be
jumping around quite a bit!

Also make sure that the delay is adjusted so the counter updates every
second, you can be updating every two seconds which nullifies the
cancellation of the noise in the L.O.

Corby Dawson


Always a good call. Click now to establish your local phone service!
http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTHbDiki6pTvGTKKCArqOqjNQCdDWO6HMJXAtNrfARisO7WahgNOso/

Stephan, I have a DMTD system that was built by NBS (now NIST) in the early 80s. The coarse phase shifter is made up of ten lengths of miniature coax cable bundled up in a shielded box with toggle switches on the front panel to select the delay used. The LSB is about 2us. It also has a fine adjust that consists of a couple varactors in a phase delay circuit. I have found that with a good quality offset L.O. I can get the precision I need (<1X10-13th at a 1 second interval) without using the fine adjust circuit. As far as the small time interval value I usually start my measurements at around 2 to 4 us delay with the DUT set for a slowly increasing phase shift, The accuracy required depends on what your measuring but with a 10Mhz input and a 1hz beat note +-1us is equal to +-1X10-13th so most counters will work just fine. Just be advised that the us and below digits will be jumping around quite a bit! Also make sure that the delay is adjusted so the counter updates every second, you can be updating every two seconds which nullifies the cancellation of the noise in the L.O. Corby Dawson ____________________________________________________________ Always a good call. Click now to establish your local phone service! http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTHbDiki6pTvGTKKCArqOqjNQCdDWO6HMJXAtNrfARisO7WahgNOso/
SS
Stephan Sandenbergh
Fri, Jul 24, 2009 6:20 PM

Corby,
Thanks - it makes sense. I was suspecting that such a low-frequency phase
shifter would be done using cable lengths. However, I can't say that I've
got much experience with this kind of thing.

I have another question:

The rule of thumb would be that the reference oscillator must be 3x more
stable than the DUT. I haven't got much to measure my standards against, but
I have got 3 identical units. According to literature the separation of
variances method provide a reasonable estimate.

How well does a measurement, based on separation of variances, compare to
one that was done using a sufficiently stable reference?

Regards,

Stephan.

2009/7/24 Corby Dawson cdelect@juno.com

Stephan,

I have a DMTD system that was built by NBS (now NIST) in the early 80s.

The coarse phase shifter is made up of ten lengths of miniature coax
cable bundled up in a shielded box with toggle switches on the front
panel to select the delay used.

The LSB is about 2us.

It also has a fine adjust that consists of a couple varactors in a phase
delay circuit.

I have found that with a good quality offset L.O. I can get the precision
I need (<1X10-13th at a 1 second interval) without using the fine adjust
circuit.

As far as the small time interval value I usually start my measurements
at around 2 to 4 us delay with the DUT set for a slowly increasing phase
shift,

The accuracy required depends on what your measuring but with a 10Mhz
input and a 1hz beat note +-1us is equal to +-1X10-13th so most counters
will work just fine. Just be advised that the us and below digits will be
jumping around quite a bit!

Also make sure that the delay is adjusted so the counter updates every
second, you can be updating every two seconds which nullifies the
cancellation of the noise in the L.O.

Corby Dawson


Always a good call. Click now to establish your local phone service!

http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTHbDiki6pTvGTKKCArqOqjNQCdDWO6HMJXAtNrfARisO7WahgNOso/


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Corby, Thanks - it makes sense. I was suspecting that such a low-frequency phase shifter would be done using cable lengths. However, I can't say that I've got much experience with this kind of thing. I have another question: The rule of thumb would be that the reference oscillator must be 3x more stable than the DUT. I haven't got much to measure my standards against, but I have got 3 identical units. According to literature the separation of variances method provide a reasonable estimate. How well does a measurement, based on separation of variances, compare to one that was done using a sufficiently stable reference? Regards, Stephan. 2009/7/24 Corby Dawson <cdelect@juno.com> > Stephan, > > I have a DMTD system that was built by NBS (now NIST) in the early 80s. > > The coarse phase shifter is made up of ten lengths of miniature coax > cable bundled up in a shielded box with toggle switches on the front > panel to select the delay used. > > The LSB is about 2us. > > It also has a fine adjust that consists of a couple varactors in a phase > delay circuit. > > I have found that with a good quality offset L.O. I can get the precision > I need (<1X10-13th at a 1 second interval) without using the fine adjust > circuit. > > As far as the small time interval value I usually start my measurements > at around 2 to 4 us delay with the DUT set for a slowly increasing phase > shift, > > The accuracy required depends on what your measuring but with a 10Mhz > input and a 1hz beat note +-1us is equal to +-1X10-13th so most counters > will work just fine. Just be advised that the us and below digits will be > jumping around quite a bit! > > Also make sure that the delay is adjusted so the counter updates every > second, you can be updating every two seconds which nullifies the > cancellation of the noise in the L.O. > > Corby Dawson > ____________________________________________________________ > Always a good call. Click now to establish your local phone service! > > http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTHbDiki6pTvGTKKCArqOqjNQCdDWO6HMJXAtNrfARisO7WahgNOso/ > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BG
Bruce Griffiths
Sat, Jul 25, 2009 12:24 AM

Stephan

For ultra fine phase shifts one can always use NIST's approach of using
a splitter (or directional coupler) to sample the OCXO signal then use a
mixer as a dc controlled attenuator, attenuate the output with a fixed
attenuator and recombine it in quadrature with the original signal. Very
small phase shifts at the reference frequency can be achieved in this way.

Stephan Sandenbergh wrote:

Corby,
Thanks - it makes sense. I was suspecting that such a low-frequency phase
shifter would be done using cable lengths. However, I can't say that I've
got much experience with this kind of thing.

I have another question:

The rule of thumb would be that the reference oscillator must be 3x more
stable than the DUT. I haven't got much to measure my standards against, but
I have got 3 identical units. According to literature the separation of
variances method provide a reasonable estimate.

How well does a measurement, based on separation of variances, compare to
one that was done using a sufficiently stable reference?

It works best when all 3 oscillators are compared (pairwise)
simultaneously using 3 DMTD systems and 3 time interval counters.
It tends to be less useful for longer tau as the 3 oscillators share the
same external environment.

Regards,

Stephan.

Bruce

2009/7/24 Corby Dawson cdelect@juno.com

Stephan,

I have a DMTD system that was built by NBS (now NIST) in the early 80s.

The coarse phase shifter is made up of ten lengths of miniature coax
cable bundled up in a shielded box with toggle switches on the front
panel to select the delay used.

The LSB is about 2us.

It also has a fine adjust that consists of a couple varactors in a phase
delay circuit.

I have found that with a good quality offset L.O. I can get the precision
I need (<1X10-13th at a 1 second interval) without using the fine adjust
circuit.

As far as the small time interval value I usually start my measurements
at around 2 to 4 us delay with the DUT set for a slowly increasing phase
shift,

The accuracy required depends on what your measuring but with a 10Mhz
input and a 1hz beat note +-1us is equal to +-1X10-13th so most counters
will work just fine. Just be advised that the us and below digits will be
jumping around quite a bit!

Also make sure that the delay is adjusted so the counter updates every
second, you can be updating every two seconds which nullifies the
cancellation of the noise in the L.O.

Corby Dawson


Always a good call. Click now to establish your local phone service!

http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTHbDiki6pTvGTKKCArqOqjNQCdDWO6HMJXAtNrfARisO7WahgNOso/


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Stephan For ultra fine phase shifts one can always use NIST's approach of using a splitter (or directional coupler) to sample the OCXO signal then use a mixer as a dc controlled attenuator, attenuate the output with a fixed attenuator and recombine it in quadrature with the original signal. Very small phase shifts at the reference frequency can be achieved in this way. Stephan Sandenbergh wrote: > Corby, > Thanks - it makes sense. I was suspecting that such a low-frequency phase > shifter would be done using cable lengths. However, I can't say that I've > got much experience with this kind of thing. > > I have another question: > > The rule of thumb would be that the reference oscillator must be 3x more > stable than the DUT. I haven't got much to measure my standards against, but > I have got 3 identical units. According to literature the separation of > variances method provide a reasonable estimate. > > How well does a measurement, based on separation of variances, compare to > one that was done using a sufficiently stable reference? > > It works best when all 3 oscillators are compared (pairwise) simultaneously using 3 DMTD systems and 3 time interval counters. It tends to be less useful for longer tau as the 3 oscillators share the same external environment. > Regards, > > Stephan. > > > Bruce > 2009/7/24 Corby Dawson <cdelect@juno.com> > > >> Stephan, >> >> I have a DMTD system that was built by NBS (now NIST) in the early 80s. >> >> The coarse phase shifter is made up of ten lengths of miniature coax >> cable bundled up in a shielded box with toggle switches on the front >> panel to select the delay used. >> >> The LSB is about 2us. >> >> It also has a fine adjust that consists of a couple varactors in a phase >> delay circuit. >> >> I have found that with a good quality offset L.O. I can get the precision >> I need (<1X10-13th at a 1 second interval) without using the fine adjust >> circuit. >> >> As far as the small time interval value I usually start my measurements >> at around 2 to 4 us delay with the DUT set for a slowly increasing phase >> shift, >> >> The accuracy required depends on what your measuring but with a 10Mhz >> input and a 1hz beat note +-1us is equal to +-1X10-13th so most counters >> will work just fine. Just be advised that the us and below digits will be >> jumping around quite a bit! >> >> Also make sure that the delay is adjusted so the counter updates every >> second, you can be updating every two seconds which nullifies the >> cancellation of the noise in the L.O. >> >> Corby Dawson >> ____________________________________________________________ >> Always a good call. Click now to establish your local phone service! >> >> http://thirdpartyoffers.juno.com/TGL2141/fc/BLSrjpTHbDiki6pTvGTKKCArqOqjNQCdDWO6HMJXAtNrfARisO7WahgNOso/ >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > >