Discussion of precise voltage measurement

View all threads

Trimming the LTZ1000 tempco

AJ
Andreas Jahn
Thu, Oct 6, 2011 9:37 PM

It seems that the page 6 trick slightly reduces the tempco but it
still stays in the 50ppm/C area. Or I have forgot something in my
calculations. I don't have a spare LTZ1000 available at the moment to
run the tests but maybe Andreas has time in the future.

Hello,

But consider: a resistance change on the 200R resistor of page 6
will directly affect output voltage. 200R * 5mA = 1V.
A 100 ppm change of the 200R Resistor has a 1V/8.2V
= 12 ppm change on output voltage.
All other resistors have below 1ppm change with 100ppm resistance change.

So the 200 Ohms resistor will be the most
critical for tempco and ageing in the whole system.

So in my opinion its only worth to consider if you
really need to save the 20-30 mA for the heater.

With best regards

Andreas

> > It seems that the page 6 trick slightly reduces the tempco but it > still stays in the 50ppm/C area. Or I have forgot something in my > calculations. I don't have a spare LTZ1000 available at the moment to > run the tests but maybe Andreas has time in the future. > Hello, But consider: a resistance change on the 200R resistor of page 6 will directly affect output voltage. 200R * 5mA = 1V. A 100 ppm change of the 200R Resistor has a 1V/8.2V = 12 ppm change on output voltage. All other resistors have below 1ppm change with 100ppm resistance change. So the 200 Ohms resistor will be the most critical for tempco and ageing in the whole system. So in my opinion its only worth to consider if you really need to save the 20-30 mA for the heater. With best regards Andreas
Михаил
Fri, Oct 7, 2011 4:16 AM

Hello,

Please, look at the actual measurement results, maked by LTZ guru Lymex
Zhang/BG2VO (GIF attached).
And some information about tempco:
http://bbs.38hot.net/read.php?tid=5569
http://bbs.38hot.net/read.php?tid=2880

With best regards
Mickle T.

Friday, October 7, 2011, 1:37:38 AM, you wrote:

It seems that the page 6 trick slightly reduces the tempco but it
still stays in the 50ppm/C area. Or I have forgot something in my
calculations. I don't have a spare LTZ1000 available at the moment to
run the tests but maybe Andreas has time in the future.

AJ> Hello,
AJ> But consider: a resistance change on the 200R resistor of page 6
AJ> will directly affect output voltage. 200R * 5mA = 1V.
AJ> A 100 ppm change of the 200R Resistor has a 1V/8.2V
AJ> = 12 ppm change on output voltage.
AJ> All other resistors have below 1ppm change with 100ppm resistance change.

AJ> So the 200 Ohms resistor will be the most
AJ> critical for tempco and ageing in the whole system.

AJ> So in my opinion its only worth to consider if you
AJ> really need to save the 20-30 mA for the heater.

AJ> With best regards
AJ> Andreas

Hello, Please, look at the actual measurement results, maked by LTZ guru Lymex Zhang/BG2VO (GIF attached). And some information about tempco: http://bbs.38hot.net/read.php?tid=5569 http://bbs.38hot.net/read.php?tid=2880 With best regards Mickle T. Friday, October 7, 2011, 1:37:38 AM, you wrote: >> >> It seems that the page 6 trick slightly reduces the tempco but it >> still stays in the 50ppm/C area. Or I have forgot something in my >> calculations. I don't have a spare LTZ1000 available at the moment to >> run the tests but maybe Andreas has time in the future. >> AJ> Hello, AJ> But consider: a resistance change on the 200R resistor of page 6 AJ> will directly affect output voltage. 200R * 5mA = 1V. AJ> A 100 ppm change of the 200R Resistor has a 1V/8.2V AJ> = 12 ppm change on output voltage. AJ> All other resistors have below 1ppm change with 100ppm resistance change. AJ> So the 200 Ohms resistor will be the most AJ> critical for tempco and ageing in the whole system. AJ> So in my opinion its only worth to consider if you AJ> really need to save the 20-30 mA for the heater. AJ> With best regards AJ> Andreas
BG
Bruce Griffiths
Fri, Oct 7, 2011 7:01 AM

Andreas Jahn wrote:

It seems that the page 6 trick slightly reduces the tempco but it
still stays in the 50ppm/C area. Or I have forgot something in my
calculations. I don't have a spare LTZ1000 available at the moment to
run the tests but maybe Andreas has time in the future.

Hello,

But consider: a resistance change on the 200R resistor of page 6
will directly affect output voltage. 200R * 5mA = 1V.
A 100 ppm change of the 200R Resistor has a 1V/8.2V = 12 ppm change on
output voltage.
All other resistors have below 1ppm change with 100ppm resistance change.

So the 200 Ohms resistor will be the most critical for tempco and
ageing in the whole system.

So in my opinion its only worth to consider if you really need to save
the 20-30 mA for the heater.

With best regards

Andreas

Except that in your case you should only need around 10% of 200 ohms
with a correspondingly reduced sensitivity to the tempco of this resistor.

Bruce

Andreas Jahn wrote: >> >> It seems that the page 6 trick slightly reduces the tempco but it >> still stays in the 50ppm/C area. Or I have forgot something in my >> calculations. I don't have a spare LTZ1000 available at the moment to >> run the tests but maybe Andreas has time in the future. >> > Hello, > > But consider: a resistance change on the 200R resistor of page 6 > will directly affect output voltage. 200R * 5mA = 1V. > A 100 ppm change of the 200R Resistor has a 1V/8.2V = 12 ppm change on > output voltage. > All other resistors have below 1ppm change with 100ppm resistance change. > > So the 200 Ohms resistor will be the most critical for tempco and > ageing in the whole system. > > So in my opinion its only worth to consider if you really need to save > the 20-30 mA for the heater. > > With best regards > > Andreas > Except that in your case you should only need around 10% of 200 ohms with a correspondingly reduced sensitivity to the tempco of this resistor. Bruce
W
Will
Fri, Oct 7, 2011 9:00 AM

I would take the output voltage between pins 3 and 7 and leave the
resistor "outside". There is already a diode between the op-amp output
and the zener cathode and the diode must be much less time and
temperature stable than a resistor.

Or did I misunderstand something again?

Will

2011/10/7, Andreas Jahn Andreas_-_Jahn@t-online.de:

But consider: a resistance change on the 200R resistor of page 6
will directly affect output voltage. 200R * 5mA = 1V.
A 100 ppm change of the 200R Resistor has a 1V/8.2V
= 12 ppm change on output voltage.
All other resistors have below 1ppm change with 100ppm resistance change.

So the 200 Ohms resistor will be the most
critical for tempco and ageing in the whole system.

So in my opinion its only worth to consider if you
really need to save the 20-30 mA for the heater.

With best regards

Andreas

I would take the output voltage between pins 3 and 7 and leave the resistor "outside". There is already a diode between the op-amp output and the zener cathode and the diode must be much less time and temperature stable than a resistor. Or did I misunderstand something again? Will 2011/10/7, Andreas Jahn <Andreas_-_Jahn@t-online.de>: > But consider: a resistance change on the 200R resistor of page 6 > will directly affect output voltage. 200R * 5mA = 1V. > A 100 ppm change of the 200R Resistor has a 1V/8.2V > = 12 ppm change on output voltage. > All other resistors have below 1ppm change with 100ppm resistance change. > > So the 200 Ohms resistor will be the most > critical for tempco and ageing in the whole system. > > So in my opinion its only worth to consider if you > really need to save the 20-30 mA for the heater. > > With best regards > > Andreas > > >
BG
Bruce Griffiths
Fri, Oct 7, 2011 9:19 AM

There's little point in adding the tempco compensation resistor if you
do that as your proposed connection essentially eliminates the
additional correction provided by the resistor.
There's no way of avoiding the sensitivity to the value of this resistor
if this method of tempco trimming is used.
However if a +50ppm/K LTZ1000 residual tempco (with the tempco trim
resistor = 0) is typical, then the sensitivity will only be about 10% or
so of that esstimated by Adreas.

Bruce

Will wrote:

I would take the output voltage between pins 3 and 7 and leave the
resistor "outside". There is already a diode between the op-amp output
and the zener cathode and the diode must be much less time and
temperature stable than a resistor.

Or did I misunderstand something again?

Will

2011/10/7, Andreas JahnAndreas_-_Jahn@t-online.de:

But consider: a resistance change on the 200R resistor of page 6
will directly affect output voltage. 200R * 5mA = 1V.
A 100 ppm change of the 200R Resistor has a 1V/8.2V
= 12 ppm change on output voltage.
All other resistors have below 1ppm change with 100ppm resistance change.

So the 200 Ohms resistor will be the most
critical for tempco and ageing in the whole system.

So in my opinion its only worth to consider if you
really need to save the 20-30 mA for the heater.

With best regards

Andreas

There's little point in adding the tempco compensation resistor if you do that as your proposed connection essentially eliminates the additional correction provided by the resistor. There's no way of avoiding the sensitivity to the value of this resistor if this method of tempco trimming is used. However if a +50ppm/K LTZ1000 residual tempco (with the tempco trim resistor = 0) is typical, then the sensitivity will only be about 10% or so of that esstimated by Adreas. Bruce Will wrote: > I would take the output voltage between pins 3 and 7 and leave the > resistor "outside". There is already a diode between the op-amp output > and the zener cathode and the diode must be much less time and > temperature stable than a resistor. > > Or did I misunderstand something again? > > Will > > > 2011/10/7, Andreas Jahn<Andreas_-_Jahn@t-online.de>: > >> But consider: a resistance change on the 200R resistor of page 6 >> will directly affect output voltage. 200R * 5mA = 1V. >> A 100 ppm change of the 200R Resistor has a 1V/8.2V >> = 12 ppm change on output voltage. >> All other resistors have below 1ppm change with 100ppm resistance change. >> >> So the 200 Ohms resistor will be the most >> critical for tempco and ageing in the whole system. >> >> So in my opinion its only worth to consider if you >> really need to save the 20-30 mA for the heater. >> >> With best regards >> >> Andreas >> >> >> > >> >
W
Will
Fri, Oct 7, 2011 10:08 AM

Thanks Bruce.

Will wrote:

Or did I misunderstand something again?

Yes I did. And I had the same mistake with my previous calculations,
which explains the strange results.

Since you are an expert, an "LTZ1000 circuit description for dummies"
would be great.

The heater control is straightforward, but all the interactions in the
reference circuit seem not to be clear to me. And most likely there
are other dummies too.

Will

2011/10/7, Bruce Griffiths bruce.griffiths@xtra.co.nz:

There's little point in adding the tempco compensation resistor if you
do that as your proposed connection essentially eliminates the
additional correction provided by the resistor.
There's no way of avoiding the sensitivity to the value of this resistor
if this method of tempco trimming is used.
However if a +50ppm/K LTZ1000 residual tempco (with the tempco trim
resistor = 0) is typical, then the sensitivity will only be about 10% or
so of that esstimated by Adreas.

Bruce

Thanks Bruce. Will wrote: > Or did I misunderstand something again? Yes I did. And I had the same mistake with my previous calculations, which explains the strange results. Since you are an expert, an "LTZ1000 circuit description for dummies" would be great. The heater control is straightforward, but all the interactions in the reference circuit seem not to be clear to me. And most likely there are other dummies too. Will 2011/10/7, Bruce Griffiths <bruce.griffiths@xtra.co.nz>: > There's little point in adding the tempco compensation resistor if you > do that as your proposed connection essentially eliminates the > additional correction provided by the resistor. > There's no way of avoiding the sensitivity to the value of this resistor > if this method of tempco trimming is used. > However if a +50ppm/K LTZ1000 residual tempco (with the tempco trim > resistor = 0) is typical, then the sensitivity will only be about 10% or > so of that esstimated by Adreas. > > Bruce >
BG
Bruce Griffiths
Fri, Oct 7, 2011 10:49 AM

In the standard positive reference circuit depicted in the latest datasheet

Vref  = Vz +Vbe
Transistor collector current, Ic = Vz/R2 (This increases slightly with
temperature due to the positive tempco of Vz)

Zener current, Iz = Vbe/R1 (This decreases with temperature due to the
negative tempco of Vbe.)

where Vz is the zener voltage, Vbe is the base to emitter voltage of the
temperature compensation transistor.

Vz has a tempco of approximately +2mV/K and Vbe has an approximate
tempco of -2mV/K
Thus Vref is nominally compensated for temperature changes.
However the tempco matching isnt perfect with a "typical" (note sample
size = 1!!!!) residual of +50ppm/K (+350uV/K)
This can be reduced by adding a resistor (Rt) in series with the zener.

In this case (to a first order approximation)

Vref = Vz + (1+ Rt/R1)*Vbe
Ic = Vz/R2 +((Rt/R1)*Vbe)/R2
Iz = Vbe/R1
If the value of Rt is suitably selected the tempco can be adjusted to zero.
However the required value will vary for each LTZ1000 and the circuit
only works if the tempco with Rt =0 is positive.
The magnitude of the Vbe tempco can be increased by reducing the
collector current (requires increasing R2), however the required
collector current to achieve zero Tc will probably be 1uA or even less,
this increases the transistor noise and makes the circuit more
susceptible to leakage shunting R2.
The compensating transistor is operated at zero collector base voltage
which eliminates the variations in Vbe due to the Early effect.

Bruce

Will wrote:

Thanks Bruce.

Will wrote:

Or did I misunderstand something again?

Yes I did. And I had the same mistake with my previous calculations,
which explains the strange results.

Since you are an expert, an "LTZ1000 circuit description for dummies"
would be great.

The heater control is straightforward, but all the interactions in the
reference circuit seem not to be clear to me. And most likely there
are other dummies too.

Will

2011/10/7, Bruce Griffithsbruce.griffiths@xtra.co.nz:

There's little point in adding the tempco compensation resistor if you
do that as your proposed connection essentially eliminates the
additional correction provided by the resistor.
There's no way of avoiding the sensitivity to the value of this resistor
if this method of tempco trimming is used.
However if a +50ppm/K LTZ1000 residual tempco (with the tempco trim
resistor = 0) is typical, then the sensitivity will only be about 10% or
so of that esstimated by Adreas.

Bruce

In the standard positive reference circuit depicted in the latest datasheet Vref = Vz +Vbe Transistor collector current, Ic = Vz/R2 (This increases slightly with temperature due to the positive tempco of Vz) Zener current, Iz = Vbe/R1 (This decreases with temperature due to the negative tempco of Vbe.) where Vz is the zener voltage, Vbe is the base to emitter voltage of the temperature compensation transistor. Vz has a tempco of approximately +2mV/K and Vbe has an approximate tempco of -2mV/K Thus Vref is nominally compensated for temperature changes. However the tempco matching isnt perfect with a "typical" (note sample size = 1!!!!) residual of +50ppm/K (+350uV/K) This can be reduced by adding a resistor (Rt) in series with the zener. In this case (to a first order approximation) Vref = Vz + (1+ Rt/R1)*Vbe Ic = Vz/R2 +((Rt/R1)*Vbe)/R2 Iz = Vbe/R1 If the value of Rt is suitably selected the tempco can be adjusted to zero. However the required value will vary for each LTZ1000 and the circuit only works if the tempco with Rt =0 is positive. The magnitude of the Vbe tempco can be increased by reducing the collector current (requires increasing R2), however the required collector current to achieve zero Tc will probably be 1uA or even less, this increases the transistor noise and makes the circuit more susceptible to leakage shunting R2. The compensating transistor is operated at zero collector base voltage which eliminates the variations in Vbe due to the Early effect. Bruce Will wrote: > Thanks Bruce. > > Will wrote: > >> Or did I misunderstand something again? >> > Yes I did. And I had the same mistake with my previous calculations, > which explains the strange results. > > Since you are an expert, an "LTZ1000 circuit description for dummies" > would be great. > > The heater control is straightforward, but all the interactions in the > reference circuit seem not to be clear to me. And most likely there > are other dummies too. > > Will > > > > 2011/10/7, Bruce Griffiths<bruce.griffiths@xtra.co.nz>: > >> There's little point in adding the tempco compensation resistor if you >> do that as your proposed connection essentially eliminates the >> additional correction provided by the resistor. >> There's no way of avoiding the sensitivity to the value of this resistor >> if this method of tempco trimming is used. >> However if a +50ppm/K LTZ1000 residual tempco (with the tempco trim >> resistor = 0) is typical, then the sensitivity will only be about 10% or >> so of that esstimated by Adreas. >> >> Bruce >> >> > >
AJ
Andreas Jahn
Fri, Oct 7, 2011 6:00 PM

However the tempco matching isnt perfect with a "typical" (note sample
size = 1!!!!) residual of +50ppm/K (+350uV/K)

I would not talk of a sample size = 1.
My LTZ #1 has  48ppm/K with 70k and 52ppm/K with 50K
My LTZ #2 has  54ppm/K with 70k
Both have a setpoint divider of 12K5/1K.

and:

from the datasheet you can calculate for the 13K/1K nominal temperature
setpoint
100 ppm in voltage divider change R4/R5 will give 1 ppm output change.
100 ppm change of 0,5V VBe will give around 50uV temperature setpoint
change.
2mV VBE/K or 50uV/2000uV/K = 0,025K temperature setpoint change.

So from datasheet with nominal divider you can calculate
40 ppm / K typical temperature gradient.

From Mickles "LTZ-guru" we can calculate 39ppm/K for the

unheated reference. (0.95ppm output for 100ppm divider change)

So there seems to be less tempco (but faster ageing)
with higher temperature setpoints.

With best regards

Andreas

> However the tempco matching isnt perfect with a "typical" (note sample > size = 1!!!!) residual of +50ppm/K (+350uV/K) I would not talk of a sample size = 1. My LTZ #1 has 48ppm/K with 70k and 52ppm/K with 50K My LTZ #2 has 54ppm/K with 70k Both have a setpoint divider of 12K5/1K. and: from the datasheet you can calculate for the 13K/1K nominal temperature setpoint 100 ppm in voltage divider change R4/R5 will give 1 ppm output change. 100 ppm change of 0,5V VBe will give around 50uV temperature setpoint change. 2mV VBE/K or 50uV/2000uV/K = 0,025K temperature setpoint change. So from datasheet with nominal divider you can calculate 40 ppm / K typical temperature gradient. >From Mickles "LTZ-guru" we can calculate 39ppm/K for the unheated reference. (0.95ppm output for 100ppm divider change) So there seems to be less tempco (but faster ageing) with higher temperature setpoints. With best regards Andreas
BG
Bruce Griffiths
Fri, Oct 7, 2011 6:40 PM

Andreas Jahn wrote:

However the tempco matching isnt perfect with a "typical" (note
sample size = 1!!!!) residual of +50ppm/K (+350uV/K)

I would not talk of a sample size = 1.

Even a sample size of 2 is too small to be useful in predicting the
tempco of the entire population of LTZ1000s
There will be variations due to manufacturing tolerances.
Perhaps your 2 LTZ1000s came from the same batch/wafer?
Measuring the tempcos of a statistically significant sample would be
more useful.

My LTZ #1 has  48ppm/K with 70k and 52ppm/K with 50K
My LTZ #2 has  54ppm/K with 70k
Both have a setpoint divider of 12K5/1K.

and:

from the datasheet you can calculate for the 13K/1K nominal
temperature setpoint
100 ppm in voltage divider change R4/R5 will give 1 ppm output change.
100 ppm change of 0,5V VBe will give around 50uV temperature setpoint
change.
2mV VBE/K or 50uV/2000uV/K = 0,025K temperature setpoint change.

So from datasheet with nominal divider you can calculate
40 ppm / K typical temperature gradient.

The major problem with that analysis is that the sign of the tempco
cannot be deduced and this only gives an approximate upper bound for the
magnitude of the tempco.
The tempco could lie anywhere within the [-40, +40] ppm/K range.

From Mickles "LTZ-guru" we can calculate 39ppm/K for the

unheated reference. (0.95ppm output for 100ppm divider change)

So there seems to be less tempco (but faster ageing)
with higher temperature setpoints.

With best regards

Andreas

Bruce

Andreas Jahn wrote: >> However the tempco matching isnt perfect with a "typical" (note >> sample size = 1!!!!) residual of +50ppm/K (+350uV/K) > > I would not talk of a sample size = 1. Even a sample size of 2 is too small to be useful in predicting the tempco of the entire population of LTZ1000s There will be variations due to manufacturing tolerances. Perhaps your 2 LTZ1000s came from the same batch/wafer? Measuring the tempcos of a statistically significant sample would be more useful. > My LTZ #1 has 48ppm/K with 70k and 52ppm/K with 50K > My LTZ #2 has 54ppm/K with 70k > Both have a setpoint divider of 12K5/1K. > > and: > > from the datasheet you can calculate for the 13K/1K nominal > temperature setpoint > 100 ppm in voltage divider change R4/R5 will give 1 ppm output change. > 100 ppm change of 0,5V VBe will give around 50uV temperature setpoint > change. > 2mV VBE/K or 50uV/2000uV/K = 0,025K temperature setpoint change. > > So from datasheet with nominal divider you can calculate > 40 ppm / K typical temperature gradient. The major problem with that analysis is that the sign of the tempco cannot be deduced and this only gives an approximate upper bound for the magnitude of the tempco. The tempco could lie anywhere within the [-40, +40] ppm/K range. > >> From Mickles "LTZ-guru" we can calculate 39ppm/K for the > unheated reference. (0.95ppm output for 100ppm divider change) > > So there seems to be less tempco (but faster ageing) > with higher temperature setpoints. > > With best regards > > Andreas > Bruce
MK
m k
Fri, Oct 7, 2011 9:15 PM

Hi,

Well that expert on 38hot also got a +40ppm tempco.

http://www.national.com/rap/Story/vbe.html
This page shows that he was right to say a much lower collector current would be required. A smaller zener current and a lower collector current will reduce a + tempco, but at the expense of higher noise, so a tradeoff is that we must increase the noise to reduce the required accuracy of the temp control loop.
It seems to me that  a lower set temp would reduce the sensitivity to orientation.

Date: Sat, 8 Oct 2011 07:40:12 +1300
From: bruce.griffiths@xtra.co.nz
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Trimming the LTZ1000 tempco

Andreas Jahn wrote:

However the tempco matching isnt perfect with a "typical" (note
sample size = 1!!!!) residual of +50ppm/K (+350uV/K)

I would not talk of a sample size = 1.

Even a sample size of 2 is too small to be useful in predicting the
tempco of the entire population of LTZ1000s
There will be variations due to manufacturing tolerances.
Perhaps your 2 LTZ1000s came from the same batch/wafer?
Measuring the tempcos of a statistically significant sample would be
more useful.

My LTZ #1 has  48ppm/K with 70k and 52ppm/K with 50K
My LTZ #2 has  54ppm/K with 70k
Both have a setpoint divider of 12K5/1K.

and:

from the datasheet you can calculate for the 13K/1K nominal
temperature setpoint
100 ppm in voltage divider change R4/R5 will give 1 ppm output change.
100 ppm change of 0,5V VBe will give around 50uV temperature setpoint
change.
2mV VBE/K or 50uV/2000uV/K = 0,025K temperature setpoint change.

So from datasheet with nominal divider you can calculate
40 ppm / K typical temperature gradient.

The major problem with that analysis is that the sign of the tempco
cannot be deduced and this only gives an approximate upper bound for the
magnitude of the tempco.
The tempco could lie anywhere within the [-40, +40] ppm/K range.

From Mickles "LTZ-guru" we can calculate 39ppm/K for the

unheated reference. (0.95ppm output for 100ppm divider change)

So there seems to be less tempco (but faster ageing)
with higher temperature setpoints.

With best regards

Andreas

Bruce


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi, Well that expert on 38hot also got a +40ppm tempco. http://www.national.com/rap/Story/vbe.html This page shows that he was right to say a much lower collector current would be required. A smaller zener current and a lower collector current will reduce a + tempco, but at the expense of higher noise, so a tradeoff is that we must increase the noise to reduce the required accuracy of the temp control loop. It seems to me that a lower set temp would reduce the sensitivity to orientation. > Date: Sat, 8 Oct 2011 07:40:12 +1300 > From: bruce.griffiths@xtra.co.nz > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Trimming the LTZ1000 tempco > > Andreas Jahn wrote: > >> However the tempco matching isnt perfect with a "typical" (note > >> sample size = 1!!!!) residual of +50ppm/K (+350uV/K) > > > > I would not talk of a sample size = 1. > Even a sample size of 2 is too small to be useful in predicting the > tempco of the entire population of LTZ1000s > There will be variations due to manufacturing tolerances. > Perhaps your 2 LTZ1000s came from the same batch/wafer? > Measuring the tempcos of a statistically significant sample would be > more useful. > > My LTZ #1 has 48ppm/K with 70k and 52ppm/K with 50K > > My LTZ #2 has 54ppm/K with 70k > > Both have a setpoint divider of 12K5/1K. > > > > and: > > > > from the datasheet you can calculate for the 13K/1K nominal > > temperature setpoint > > 100 ppm in voltage divider change R4/R5 will give 1 ppm output change. > > 100 ppm change of 0,5V VBe will give around 50uV temperature setpoint > > change. > > 2mV VBE/K or 50uV/2000uV/K = 0,025K temperature setpoint change. > > > > So from datasheet with nominal divider you can calculate > > 40 ppm / K typical temperature gradient. > The major problem with that analysis is that the sign of the tempco > cannot be deduced and this only gives an approximate upper bound for the > magnitude of the tempco. > The tempco could lie anywhere within the [-40, +40] ppm/K range. > > > > >> From Mickles "LTZ-guru" we can calculate 39ppm/K for the > > unheated reference. (0.95ppm output for 100ppm divider change) > > > > So there seems to be less tempco (but faster ageing) > > with higher temperature setpoints. > > > > With best regards > > > > Andreas > > > Bruce > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.