Some questions to zeners (1N823-1N829)

W
WarrenS
Sun, Jan 27, 2013 10:23 PM

Andreas

1st my two cents worth concerning comments of others that are being more
practical than nutty.

"it is not easy to achieve these specs",
Too True,
If you do not already know this, then I'd suggest you back off and do some
major readjustment of your requirements.
Assuming you already know it is not easy and why, then
IMHO, what you want to make is doable with the right compromises, some of
which being
A little extra power.
a lot of extra time
a lot of test selecting
some extra circuits or S/W

"your expectations are not realistic "
True,  if you think it can be done from a data sheet and a few of off the
shelf parts in a production device.
Not true, if you want to go totally nuts and have the time to do it.

"The thermoelectric effects causing error would swamp the performance"
Could be true for some. If you do not already know enough to keep these
effects well under <<0.1PPM then should find another starter project.

"To keep everything below the 1 ppm/deg C range you would have to put the
entire circuit in controlled temperature"
Too true, But for these requirements, fortunately you do not have to keep
Everything under 1 ppm/deg C
ALL you have to do is to keep the sum total of Everything under 1ppm.  Big
difference!
And that is very easy to do, by a fact of ten plus if desired.

"long term drift and noise will be intrinsic to the devices, and
unpredictable except in a statistical sense."
Half true,
Using the right device, and a lot of time, noise can be low enough to be
mostly insignificant
and long term drift will follow a predicable slope.
It ain't easy and it's going to take a lot of time. If you think otherwise,
you need a new project..

"provide near zero tempco at one temperature only"
Answer too limited to be useful
zero TCing a circuit, can provide no voltage difference at any two
temperatures, (to the limit of the repeatability)
Depending how flat you need it between these two temperatures, is the only
issue.
Best to plan on having a second order  TC method as well.
A total max total deviation of 1ppm and a 0.1 PPM /deg is not too hard to
get.

"One of the best voltage standard Datron 4910AV (4x LTZ1000) have only 1
ppm/year drift"
OK, so that shows there is at least one way to do it, so it is not
impossible.
Now just need to find the best way for you to do it.

"Don't bother with TC zeners"
Not a bad idea, but Unfortunately Not a lot of other choices considering
your requirements.
The other choice you have (that you should consider if time is a big issue)
is to get some three terminal Fluke voltage references.
If necessary by removing them from old test equipment such as Fluke 731
Those have already been selected and aged, and you'd just then need to work
on the long term drift compensation selection method.

"There are lots of nice IC references available"
True, and if you can live with data sheet specs, they are much better
choices.
But hard if not impossible to find anything that will compare to 1/F pop
corn noise, and long term stability

"I doubt that any TCZ will match an LM399"
True when considering a wider temperature range, and it sure makes things
easier,
so a good suggestion if TC was your main problem.
But TC need not be a problem, and most any good zener will outperform most
any selected LM399 in low freq noise and stability.

"so would have to be ovenized to get best performance".
True of Any circuit when "best" TC is concerned, so not very relevant,
The question is can it be made good enough without an oven?
And the answer is defiantly yes with the right tempco circuit.
And if you want it even better, can make a very low power mini-oven.

Back to your email ******************

Do you have typical values over a 64-90 ?F range. Will it be above 1ppm/K
or  below?

I do not have any data over that Temp range in front of me.
Over My room changes (about 1/2 that temp range), can keep the Total voltage
error down to around 1PPM, (not just /deg)
If it is an important consideration, I may try and see what happens over a
wider range.
For a backup plan, considering having an addition second order TC
compensating method such as S/W.
Overall compensating to 1ppm / C is no problem,
which is Total change +- 7PPM over a +- 7 deg C range
Ten times better is possible over narrow temp ranges like that.

It is the hysteresis and stability that is going to be a limiting factor,
and last I looked,
The zeners I tested had no measurable Hysteresis over a much wider temp
range than that.

From your plot it would be 0.33ppm per 3 degrees in narrow range i.e. 0.1
ppm per degree.

Correct, but not a relevant measurement.
ANY straight line TC as seen there can be zeroed out.
That was just a pre-test plot to see if the part was low enough noise to do
a more accurate Zero TC.

By the way: is it degrees Fahrenheit or degrees Celsius (= 3 Kelvin)?

My world is mostly in F, and since I was not even plotting temp at the same
time, pretty irrelevant for that data set.

If I understand you right then you would not use this device because it
does
behave other than the others?

Your understaning is not correct, not even close. My comment applies to a
preselection process of 1N82x parts.
In the case you stated, your preselection process (assuming you have enough
to pick from) would be to select all the parts that are less than say 1PPM,
with maybe a 20% yield.

If the yield is too low then make a second pass for parts that are say 2 +-1
PPM,
and in that case, then yes the "best" part would not go into that batch,
because it will likely need a different compensation than the others.

On the other side it seems to be the device with the largest ageing rate
of  the 5 pieces.

likely too little data to be important, Just as likely random luck as
anything important.
Until after you do some pre-aging, (whatever that may mean for that part) I
would not even bother looking at that data this early on in the selection
process.

So I still hope that anyone has experiences with hysteresis of the zeners.

Turn on after power down, and hysteresis, repeatable, etc altogether is
under 1PPM.
As part of my pre-test TC procedure, I hit them with cold spray to 0 C and
heat them with a heat gun to ~50C, a few times,
If they are not repeatable they are not used in further testing.

I have a question about two of your requirements
AJ> tempco below 1ppm/K
AJ> hysteresis in the 10-40 degree range well below 1ppm

This suggest to me that what you are really planning on making is something
with an overall compensted TC that is well below a 1 ppm/K,
otherwise your hysteresis requirement does not make a lot of sense to me.
So if what you are really after is more like < 1 PPM total error over time
and temp, all this extra trouble now makes a lot more sense.

Now the disclaimer, I have no idea if any of the now available 1N825 or
better parts work like I've described.
It is VERY much a manufacture sensitive thing.
Short term noise being the biggest rejecting thing I've seen in the past.
0.5 PPM (3uv steps) are not untypical for some batches.

If someone that has enough of these parts from a single batch and
manufacture to make it worth while testing them, and will make them
available to others, then I'll test and report the results and compare them
to what I have.

There may be another less direct way to get where you want to go (You did
not say where that was)
Divide the task into two or more sections.
As an example, One being the low power short term stable DVM device that
works over a limited temperature.
There would be no problem getting 0.1 PPM accuracy/repeatable for that part.
and a second higher power device that is mostly off, and powered up Just
long enough to calibrate the low power xfer DVM above.

And if you can then include some sort of cal exchange program for the higher
powered but less used device, this would make a very hard project into a
relative easy project.

ws


Subject: Re: [volt-nuts] Some questions to zeners (1N823-1N829)

Hello Warren,

many thanks for your valuable response.

If you want to make something in volume

The planned number of devices is more in the range of 2 up to four.

one of only a very few possible solutions.

I guess that the others are a LM399 or a LTZ1000 based solution (both with
heaters).

Don't just solder the parts down on a PCB without a stress relieve loop
in
the leads.

Thats a good hint otherwise I will get the humidity changes of the PCB as
stress on the device.

The zero TC current can be set so that the voltage at most any two
temperatures will be the same. (<< 1PPM)

Do you have typical values over a 64-90 ?F range. Will it be above 1ppm/K
or
below?

From your plot it would be 0.33ppm per 3 degrees in narrow range i.e. 0.1

ppm per degree.
By the way: is it degrees Fahrenheit or degrees Celsius (= 3 Kelvin)?

From a given batch, any that are considerable different, I don't use
because they may have something else wrong going on.

Thats interesting since I have 5 pieces of the brand new LT1236AILS-5
devices.
4 of them have a tempco of  2-3 ppm/K around room temperature.
1 piece has a very flat tempco of around 0.2ppm/K (picture attached)
If I understand you right then you would not use this device because it
does
behave other than the others?
On the other side it seems to be the device with the largest ageing rate
of
the 5 pieces.

One experience that I did is that devices out of one batch which have a
low
tempco around room temperature tend to have a larger hysteresis and vice
versa.

So I still hope that anyone has experiences with hysteresis of the zeners.

With best regards

Andreas


Desired requirements from Andreas:

AJ> I want to have high end specs with the power consumption
AJ> of a battery supplied DMM.
AJ> The planned number of devices is more in the range of 2 up to four.
AJ> that is:
AJ> tempco below 1ppm/K
AJ> hysteresis in the 10-40 degree range well below 1ppm
AJ> no sensitivity to humidity (so all plastic housings will fall out)
AJ> ageing in the range of 1 ppm/year after some pre-ageing


Andreas

If you want to make something in volume which is just pretty good, I
would
not recommend this method for a new design.
On the other hand, if this is a nuts thing to make the very best, this is
one of only a very few possible solutions.

Here are some general answers, most of my experience with these parts is
pretty dated, (i.e long ago).
As when pushing the performance limit of any reference, there is a  lot
of
variation between parts and even more so between manufactures.
Solution is, select, Age, select, test, grade, select.

As far as my experience with 1N823, performance depends on the run and
what
is left after the manufacture has selected out the others.
With 1N823's, Yield can often go to zero. With 1N825's a typical yield
I've
seen is around 25-50% (from the right manufacture and batch)

Yes the main difference is the zero TC current, with some parts there is
no
zero TC current.
So yes you are more likely to get a lower current TC such as 5 ma from a
1N823 or 1N825 than a 1N829, but it could of course be > than 7.5 ma.
I don't use anything that does not have a zero TC between ~ 4 and 10 ma
I found TC to be very much a Batch thing, with up to 50% of the majority
of
a batch, tending to be similar.
From a given batch, any that are considerable different, I don't use
because
they may have something else wrong going on.

Another thing that needs to be selected for in high end references, and
will
vary by manufacturer, is 1/F noise.  The random jumps in the voltage.

For me, hysteresis has not been a issue over room temperature changes for
the most part, but something that has to be checked.
Some Manufactures are better than others, and hysteresis can and will be
effected by assembly, layout, or anything that puts any stress on the
part.
Don't just solder the parts down on a PCB without a stress relieve loop
in
the leads.

The zero TC current can be set so that the voltage at most any two
temperatures will be the same. (<< 1PPM)
If the voltage change in-between those two temperatures is too much, lots
of
ways to add an additional second order temperature compensation.

For the best TC performance, consider the mini-oven idea with the zener,
heater resistor, and thermistor all  heat shrunk together.
With a lot of outer insulation, it could be done low power by adding an
addition 0 to 5 ma to the heater resistor.

Everything has it's trade offs.
The trade off using these zeners is time and complexity.
For these parts, 5ma is about as low as you are going to get.
For low power, there are many things Much better.
The trade off you make to get low power is "Noise" & stability.
The trade off you make for the good TC of LM399 is long term stability,
PPM
noise, and the high cost of selection fall outs.

The best solution will depend on many things including the desired
performance, how many you want to make, the cost you place on selection
time,
and if you can still find the 1N825's at a reasonable price like they
where
in the 70s & 80s. (&  $0.10 in 2000s)

For a xfer standard, the Most important criteria is 1/f random noise.
Most
everything else can be compensated out.
For that, it is hard to build anything better than using a 1N825 selected
device.  Plot attached.

ws


From: "Andreas Jahn" Andreas_-_Jahn@t-online.de
Subject: [volt-nuts] Some questions to zeners (1N823-1N829)

Hello all,
Hello Warren,

after having experimented a lot with 5V monolithic zener references
and still not found the ideal solution I want to try a 1N82x based
solution.

For me a extended room temperature range of
25 degrees centigrade +/- 7 degrees (64-90 ?F)  is of interest.
Since I plan to have battery supplied instruments a lower supply current
would be of interest.

For the zeners a zero TC current is stated.
Over which temperature range the TC is nearly zero.
How large is the voltage deviation in the above mentioned range?

Does it play a role for the absolute temperature deviation if a 1N823 or
a
1N829 is used?
Or is the behaviour equally when the individual zero TC current is used?
Is the only difference between the selections that the zero TC current
is
more near the 7.5mA value on the 1N829?

So is it more likely to get a low zero TC current of 4 mA on a 1N823
device than on the 1N829?
Or should I go for the 1N829A for the lowest absolute TC?

How large is the hysteresis on the zeners in a temperature range of
10-40
degrees celsius (50-104?F).
On monolithic unheated reference voltages with hermetic case I have up
to
around 2 ppm hysteresis difference
on temperature cycling. (see attached picture with 10-45 degrees celsius
on X-Axis for an ADC with a 5V reference
measuring a LM399 heated reference over a 2:1 precision voltage divider.
The ADC with the 5V reference is temperature cycled).

I blame the temperature hysteresis on the die attach to the lead frame
which seems to be usually a silver filled epoxy compound.
I hope that the hysteresis on a discrete zener is much lower.

With best regards

Andreas

Andreas 1st my two cents worth concerning comments of others that are being more practical than nutty. "it is not easy to achieve these specs", Too True, If you do not already know this, then I'd suggest you back off and do some major readjustment of your requirements. Assuming you already know it is not easy and why, then IMHO, what you want to make is doable with the right compromises, some of which being A little extra power. a lot of extra time a lot of test selecting some extra circuits or S/W "your expectations are not realistic " True, if you think it can be done from a data sheet and a few of off the shelf parts in a production device. Not true, if you want to go totally nuts and have the time to do it. "The thermoelectric effects causing error would swamp the performance" Could be true for some. If you do not already know enough to keep these effects well under <<0.1PPM then should find another starter project. "To keep everything below the 1 ppm/deg C range you would have to put the entire circuit in controlled temperature" Too true, But for these requirements, fortunately you do not have to keep *Everything* under 1 ppm/deg C ALL you have to do is to keep the sum total of Everything under 1ppm. Big difference! And that is very easy to do, by a fact of ten plus if desired. "long term drift and noise will be intrinsic to the devices, and unpredictable except in a statistical sense." Half true, Using the *right device*, and a lot of time, noise can be low enough to be mostly insignificant and long term drift will follow a predicable slope. It ain't easy and it's going to take a lot of time. If you think otherwise, you need a new project.. "provide near zero tempco at one temperature only" Answer too limited to be useful zero TCing a circuit, can provide no voltage difference at any two temperatures, (to the limit of the repeatability) Depending how flat you need it between these two temperatures, is the only issue. Best to plan on having a second order TC method as well. A total max total deviation of 1ppm and a 0.1 PPM /deg is not too hard to get. "One of the best voltage standard Datron 4910AV (4x LTZ1000) have only 1 ppm/year drift" OK, so that shows there is at least one way to do it, so it is not impossible. Now just need to find the best way for you to do it. "Don't bother with TC zeners" Not a bad idea, but Unfortunately Not a lot of other choices considering your requirements. The other choice you have (that you should consider if time is a big issue) is to get some three terminal Fluke voltage references. If necessary by removing them from old test equipment such as Fluke 731 Those have already been selected and aged, and you'd just then need to work on the long term drift compensation selection method. "There are lots of nice IC references available" True, and if you can live with data sheet specs, they are much better choices. But hard if not impossible to find anything that will compare to 1/F pop corn noise, and long term stability "I doubt that any TCZ will match an LM399" True when considering a wider temperature range, and it sure makes things easier, so a good suggestion if TC was your main problem. But TC need not be a problem, and most any good zener will outperform most any selected LM399 in low freq noise and stability. "so would have to be ovenized to get *best* performance". True of Any circuit when "best" TC is concerned, so not very relevant, The question is can it be made good enough without an oven? And the answer is defiantly yes with the right tempco circuit. And if you want it even better, can make a very low power mini-oven. Back to your email ****************** > Do you have typical values over a 64-90 ?F range. Will it be above 1ppm/K > or below? I do not have any data over that Temp range in front of me. Over My room changes (about 1/2 that temp range), can keep the Total voltage error down to around 1PPM, (not just /deg) If it is an important consideration, I may try and see what happens over a wider range. For a backup plan, considering having an addition second order TC compensating method such as S/W. Overall compensating to 1ppm / C is no problem, which is Total change +- 7PPM over a +- 7 deg C range Ten times better is possible over narrow temp ranges like that. It is the hysteresis and stability that is going to be a limiting factor, and last I looked, The zeners I tested had no measurable Hysteresis over a much wider temp range than that. >>From your plot it would be 0.33ppm per 3 degrees in narrow range i.e. 0.1 >>ppm per degree. Correct, but not a relevant measurement. ANY straight line TC as seen there can be zeroed out. That was just a pre-test plot to see if the part was low enough noise to do a more accurate Zero TC. > By the way: is it degrees Fahrenheit or degrees Celsius (= 3 Kelvin)? My world is mostly in F, and since I was not even plotting temp at the same time, pretty irrelevant for that data set. > If I understand you right then you would not use this device because it > does > behave other than the others? Your understaning is not correct, not even close. My comment applies to a preselection process of 1N82x parts. In the case you stated, your preselection process (assuming you have enough to pick from) would be to select all the parts that are less than say 1PPM, with maybe a 20% yield. If the yield is too low then make a second pass for parts that are say 2 +-1 PPM, and in that case, then yes the "best" part would not go into that batch, because it will likely need a different compensation than the others. > On the other side it seems to be the device with the largest ageing rate > of the 5 pieces. likely too little data to be important, Just as likely random luck as anything important. Until after you do some pre-aging, (whatever that may mean for that part) I would not even bother looking at that data this early on in the selection process. > So I still hope that anyone has experiences with hysteresis of the zeners. Turn on after power down, and hysteresis, repeatable, etc altogether is under 1PPM. As part of my pre-test TC procedure, I hit them with cold spray to 0 C and heat them with a heat gun to ~50C, a few times, If they are not repeatable they are not used in further testing. I have a question about two of your requirements AJ> tempco below 1ppm/K AJ> hysteresis in the 10-40 degree range well below 1ppm This suggest to me that what you are really planning on making is something with an overall compensted TC that is well below a 1 ppm/K, otherwise your hysteresis requirement does not make a lot of sense to me. So if what you are really after is more like < 1 PPM total error over time and temp, all this extra trouble now makes a lot more sense. Now the disclaimer, I have no idea if any of the now available 1N825 or better parts work like I've described. It is VERY much a manufacture sensitive thing. Short term noise being the biggest rejecting thing I've seen in the past. 0.5 PPM (3uv steps) are not untypical for some batches. If someone that has enough of these parts from a single batch and manufacture to make it worth while testing them, and will make them available to others, then I'll test and report the results and compare them to what I have. There may be another less direct way to get where you want to go (You did not say where that was) Divide the task into two or more sections. As an example, One being the low power short term stable DVM device that works over a limited temperature. There would be no problem getting 0.1 PPM accuracy/repeatable for that part. and a second higher power device that is mostly off, and powered up Just long enough to calibrate the low power xfer DVM above. And if you can then include some sort of cal exchange program for the higher powered but less used device, this would make a very hard project into a relative easy project. ws > ---------------------------------------------------------------------- > > Subject: Re: [volt-nuts] Some questions to zeners (1N823-1N829) > > Hello Warren, > > many thanks for your valuable response. > >> If you want to make something in volume > The planned number of devices is more in the range of 2 up to four. > >> one of only a very few possible solutions. > I guess that the others are a LM399 or a LTZ1000 based solution (both with > heaters). > >> Don't just solder the parts down on a PCB without a stress relieve loop >> in >> the leads. > Thats a good hint otherwise I will get the humidity changes of the PCB as > stress on the device. > >> The zero TC current can be set so that the voltage at most any two >> temperatures will be the same. (<< 1PPM) > Do you have typical values over a 64-90 ?F range. Will it be above 1ppm/K > or > below? >>From your plot it would be 0.33ppm per 3 degrees in narrow range i.e. 0.1 > ppm per degree. > By the way: is it degrees Fahrenheit or degrees Celsius (= 3 Kelvin)? > >> From a given batch, any that are considerable different, I don't use >> because they may have something else wrong going on. > Thats interesting since I have 5 pieces of the brand new LT1236AILS-5 > devices. > 4 of them have a tempco of 2-3 ppm/K around room temperature. > 1 piece has a very flat tempco of around 0.2ppm/K (picture attached) > If I understand you right then you would not use this device because it > does > behave other than the others? > On the other side it seems to be the device with the largest ageing rate > of > the 5 pieces. > > One experience that I did is that devices out of one batch which have a > low > tempco around room temperature tend to have a larger hysteresis and vice > versa. > > So I still hope that anyone has experiences with hysteresis of the zeners. > > With best regards > > Andreas > > ***************************************** Desired requirements from Andreas: AJ> I want to have high end specs with the power consumption AJ> of a battery supplied DMM. AJ> The planned number of devices is more in the range of 2 up to four. AJ> that is: AJ> tempco below 1ppm/K AJ> hysteresis in the 10-40 degree range well below 1ppm AJ> no sensitivity to humidity (so all plastic housings will fall out) AJ> ageing in the range of 1 ppm/year after some pre-ageing ************************* > >> Andreas >> >> If you want to make something in volume which is just pretty good, I >> would >> not recommend this method for a new design. >> On the other hand, if this is a nuts thing to make the very best, this is >> one of only a very few possible solutions. >> >> Here are some general answers, most of my experience with these parts is >> pretty dated, (i.e long ago). >> As when pushing the performance limit of any reference, there is a lot >> of >> variation between parts and even more so between manufactures. >> Solution is, select, Age, select, test, grade, select. >> >> As far as my experience with 1N823, performance depends on the run and >> what >> is left after the manufacture has selected out the others. >> With 1N823's, Yield can often go to zero. With 1N825's a typical yield >> I've >> seen is around 25-50% (from the right manufacture and batch) >> >> Yes the main difference is the zero TC current, with some parts there is >> no >> zero TC current. >> So yes you are more likely to get a lower current TC such as 5 ma from a >> 1N823 or 1N825 than a 1N829, but it could of course be > than 7.5 ma. >> I don't use anything that does not have a zero TC between ~ 4 and 10 ma >> I found TC to be very much a Batch thing, with up to 50% of the majority >> of >> a batch, tending to be similar. >> From a given batch, any that are considerable different, I don't use >> because >> they may have something else wrong going on. >> >> Another thing that needs to be selected for in high end references, and >> will >> vary by manufacturer, is 1/F noise. The random jumps in the voltage. >> >> For me, hysteresis has not been a issue over room temperature changes for >> the most part, but something that has to be checked. >> Some Manufactures are better than others, and hysteresis can and will be >> effected by assembly, layout, or anything that puts any stress on the >> part. >> Don't just solder the parts down on a PCB without a stress relieve loop >> in >> the leads. >> >> The zero TC current can be set so that the voltage at most any two >> temperatures will be the same. (<< 1PPM) >> If the voltage change in-between those two temperatures is too much, lots >> of >> ways to add an additional second order temperature compensation. >> >> For the best TC performance, consider the mini-oven idea with the zener, >> heater resistor, and thermistor all heat shrunk together. >> With a lot of outer insulation, it could be done low power by adding an >> addition 0 to 5 ma to the heater resistor. >> >> Everything has it's trade offs. >> The trade off using these zeners is time and complexity. >> For these parts, 5ma is about as low as you are going to get. >> For low power, there are many things Much better. >> The trade off you make to get low power is "Noise" & stability. >> The trade off you make for the good TC of LM399 is long term stability, >> PPM >> noise, and the high cost of selection fall outs. >> >> The best solution will depend on many things including the desired >> performance, how many you want to make, the cost you place on selection >> time, >> and if you can still find the 1N825's at a reasonable price like they >> where >> in the 70s & 80s. (& $0.10 in 2000s) >> >> For a xfer standard, the Most important criteria is 1/f random noise. >> Most >> everything else can be compensated out. >> For that, it is hard to build anything better than using a 1N825 selected >> device. Plot attached. >> >> >> ws >> >> ************************************ >>> From: "Andreas Jahn" <Andreas_-_Jahn@t-online.de> >>> Subject: [volt-nuts] Some questions to zeners (1N823-1N829) >>> >>> Hello all, >>> Hello Warren, >>> >>> after having experimented a lot with 5V monolithic zener references >>> and still not found the ideal solution I want to try a 1N82x based >>> solution. >>> >>> For me a extended room temperature range of >>> 25 degrees centigrade +/- 7 degrees (64-90 ?F) is of interest. >>> Since I plan to have battery supplied instruments a lower supply current >>> would be of interest. >>> >>> For the zeners a zero TC current is stated. >>> Over which temperature range the TC is nearly zero. >>> How large is the voltage deviation in the above mentioned range? >>> >>> Does it play a role for the absolute temperature deviation if a 1N823 or >>> a >>> 1N829 is used? >>> Or is the behaviour equally when the individual zero TC current is used? >>> Is the only difference between the selections that the zero TC current >>> is >>> more near the 7.5mA value on the 1N829? >>> >>> So is it more likely to get a low zero TC current of 4 mA on a 1N823 >>> device than on the 1N829? >>> Or should I go for the 1N829A for the lowest absolute TC? >>> >>> How large is the hysteresis on the zeners in a temperature range of >>> 10-40 >>> degrees celsius (50-104?F). >>> On monolithic unheated reference voltages with hermetic case I have up >>> to >>> around 2 ppm hysteresis difference >>> on temperature cycling. (see attached picture with 10-45 degrees celsius >>> on X-Axis for an ADC with a 5V reference >>> measuring a LM399 heated reference over a 2:1 precision voltage divider. >>> The ADC with the 5V reference is temperature cycled). >>> >>> I blame the temperature hysteresis on the die attach to the lead frame >>> which seems to be usually a silver filled epoxy compound. >>> I hope that the hysteresis on a discrete zener is much lower. >>> >>> With best regards >>> >>> Andreas
AJ
Andreas Jahn
Mon, Jan 28, 2013 10:47 PM

Hello Warren  and Volt Nuts:

for the first: from the experiences of Warren my project seems to be
feasible.
Im working now since 2008 on my "project".
Eliminating step by step the drawbacks of stability.
And learning a lot of real and not ideal parts.

I changed from SMD references with large hysteresis and bad stability to
plastic DIP devices being about a factor 3 better than SMD.
Finally finding out that some buried zener devices have relative low
hysteresis.
But air humidity was a large issue preventing stability below 10ppm over a
year.
So finally I try to use devices in hermetically tight cases which
eliminates humidity if carefully decoupled from PCB stress.

So what is feasible with selected parts and a 3rd order temperature
compensation is shown in the attached picture. (ADC13_TC)
A AD586LJ device as reference within ADC#13 is measuring a heated LM399
reference via a 2:1 capacitive divider.
X-Axis is temperature normalized to 25 degrees celsius -10/+15 degrees
measured directly at the AD586 reference.
Y-Axis on the left is LM399 measurement value in mV (divided by 2).
The red line is the measurement value of the ADC. The blue line the
resulting correction with a 3rd order polynominal.
Y-Axis on the right side is together with the green line the resulting
deviation in uV after calculating out the tempco of the reference.
Deviation is about +/-2 uV referenced to 3.4V which gives below +/-1ppm
resulting tempco after correction.
Unfortunately the slope of the uncompensated TC is relative large (about
1.1ppm). Together with my temperature resolution of about
0.1K/step this will give a relative large temperature step noise of up to
0.6uV/step in the 5V-Range of the ADC.
But anyway ADC #13 was the first ADC which I can use for ageing
measurements.

After a run in phase of nearly 1 year the ageing of ADC #13 stabilized.
Currently I compare ADC13 nearly every day with 3 heated references (1 LM399
= LM_2 and 2 LTZ1000A = LTZ_1/2).
The last half year the ageing is about 0.5 to 1.5 ppm for 6 months compared
to the heated references.
See picture ADC13_longterm:
X-Axis is day
Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1, blue =
LTZ1000A #2
Y-axis right is temperature in degree celsius of the temperature sensor near
ADC13 reference.

By the way: up to now I could not measure any effect which is related to
thermocouples.
Ok my temperature step noise is still too high. And probably I am using the
wrong connectors in my tests:
cheap D-Sub connectors where a metal shield is equalizing the temperature of
2 relative close neighboured contacts.

With best regards

Andreas

----- Original Message -----
From: "WarrenS" warrensjmail-one@yahoo.com
To: volt-nuts@febo.com
Sent: Sunday, January 27, 2013 11:23 PM
Subject: [volt-nuts] Some questions to zeners (1N823-1N829)

Andreas

1st my two cents worth concerning comments of others that are being more
practical than nutty.

"it is not easy to achieve these specs",
Too True,
If you do not already know this, then I'd suggest you back off and do some
major readjustment of your requirements.
Assuming you already know it is not easy and why, then
IMHO, what you want to make is doable with the right compromises, some of
which being
A little extra power.
a lot of extra time
a lot of test selecting
some extra circuits or S/W

"your expectations are not realistic "
True,  if you think it can be done from a data sheet and a few of off the
shelf parts in a production device.
Not true, if you want to go totally nuts and have the time to do it.

"The thermoelectric effects causing error would swamp the performance"
Could be true for some. If you do not already know enough to keep these
effects well under <<0.1PPM then should find another starter project.

"To keep everything below the 1 ppm/deg C range you would have to put the
entire circuit in controlled temperature"
Too true, But for these requirements, fortunately you do not have to keep
Everything under 1 ppm/deg C
ALL you have to do is to keep the sum total of Everything under 1ppm.  Big
difference!
And that is very easy to do, by a fact of ten plus if desired.

"long term drift and noise will be intrinsic to the devices, and
unpredictable except in a statistical sense."
Half true,
Using the right device, and a lot of time, noise can be low enough to be
mostly insignificant
and long term drift will follow a predicable slope.
It ain't easy and it's going to take a lot of time. If you think
otherwise, you need a new project..

"provide near zero tempco at one temperature only"
Answer too limited to be useful
zero TCing a circuit, can provide no voltage difference at any two
temperatures, (to the limit of the repeatability)
Depending how flat you need it between these two temperatures, is the only
issue.
Best to plan on having a second order  TC method as well.
A total max total deviation of 1ppm and a 0.1 PPM /deg is not too hard to
get.

"One of the best voltage standard Datron 4910AV (4x LTZ1000) have only 1
ppm/year drift"
OK, so that shows there is at least one way to do it, so it is not
impossible.
Now just need to find the best way for you to do it.

"Don't bother with TC zeners"
Not a bad idea, but Unfortunately Not a lot of other choices considering
your requirements.
The other choice you have (that you should consider if time is a big
issue)
is to get some three terminal Fluke voltage references.
If necessary by removing them from old test equipment such as Fluke 731
Those have already been selected and aged, and you'd just then need to
work on the long term drift compensation selection method.

"There are lots of nice IC references available"
True, and if you can live with data sheet specs, they are much better
choices.
But hard if not impossible to find anything that will compare to 1/F pop
corn noise, and long term stability

"I doubt that any TCZ will match an LM399"
True when considering a wider temperature range, and it sure makes things
easier,
so a good suggestion if TC was your main problem.
But TC need not be a problem, and most any good zener will outperform most
any selected LM399 in low freq noise and stability.

"so would have to be ovenized to get best performance".
True of Any circuit when "best" TC is concerned, so not very relevant,
The question is can it be made good enough without an oven?
And the answer is defiantly yes with the right tempco circuit.
And if you want it even better, can make a very low power mini-oven.

Back to your email ******************

Do you have typical values over a 64-90 ?F range. Will it be above 1ppm/K
or  below?

I do not have any data over that Temp range in front of me.
Over My room changes (about 1/2 that temp range), can keep the Total
voltage error down to around 1PPM, (not just /deg)
If it is an important consideration, I may try and see what happens over a
wider range.
For a backup plan, considering having an addition second order TC
compensating method such as S/W.
Overall compensating to 1ppm / C is no problem,
which is Total change +- 7PPM over a +- 7 deg C range
Ten times better is possible over narrow temp ranges like that.

It is the hysteresis and stability that is going to be a limiting factor,
and last I looked,
The zeners I tested had no measurable Hysteresis over a much wider temp
range than that.

From your plot it would be 0.33ppm per 3 degrees in narrow range i.e. 0.1
ppm per degree.

Correct, but not a relevant measurement.
ANY straight line TC as seen there can be zeroed out.
That was just a pre-test plot to see if the part was low enough noise to
do a more accurate Zero TC.

By the way: is it degrees Fahrenheit or degrees Celsius (= 3 Kelvin)?

My world is mostly in F, and since I was not even plotting temp at the
same time, pretty irrelevant for that data set.

If I understand you right then you would not use this device because it
does
behave other than the others?

Your understaning is not correct, not even close. My comment applies to a
preselection process of 1N82x parts.
In the case you stated, your preselection process (assuming you have
enough to pick from) would be to select all the parts that are less than
say 1PPM, with maybe a 20% yield.

If the yield is too low then make a second pass for parts that are say 2
+-1 PPM,
and in that case, then yes the "best" part would not go into that batch,
because it will likely need a different compensation than the others.

On the other side it seems to be the device with the largest ageing rate
of  the 5 pieces.

likely too little data to be important, Just as likely random luck as
anything important.
Until after you do some pre-aging, (whatever that may mean for that part)
I would not even bother looking at that data this early on in the
selection process.

So I still hope that anyone has experiences with hysteresis of the
zeners.

Turn on after power down, and hysteresis, repeatable, etc altogether is
under 1PPM.
As part of my pre-test TC procedure, I hit them with cold spray to 0 C and
heat them with a heat gun to ~50C, a few times,
If they are not repeatable they are not used in further testing.

I have a question about two of your requirements
AJ> tempco below 1ppm/K
AJ> hysteresis in the 10-40 degree range well below 1ppm

This suggest to me that what you are really planning on making is
something with an overall compensted TC that is well below a 1 ppm/K,
otherwise your hysteresis requirement does not make a lot of sense to me.
So if what you are really after is more like < 1 PPM total error over time
and temp, all this extra trouble now makes a lot more sense.

Now the disclaimer, I have no idea if any of the now available 1N825 or
better parts work like I've described.
It is VERY much a manufacture sensitive thing.
Short term noise being the biggest rejecting thing I've seen in the past.
0.5 PPM (3uv steps) are not untypical for some batches.

If someone that has enough of these parts from a single batch and
manufacture to make it worth while testing them, and will make them
available to others, then I'll test and report the results and compare
them to what I have.

There may be another less direct way to get where you want to go (You did
not say where that was)
Divide the task into two or more sections.
As an example, One being the low power short term stable DVM device that
works over a limited temperature.
There would be no problem getting 0.1 PPM accuracy/repeatable for that
part.
and a second higher power device that is mostly off, and powered up Just
long enough to calibrate the low power xfer DVM above.

And if you can then include some sort of cal exchange program for the
higher powered but less used device, this would make a very hard project
into a relative easy project.

ws


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hello Warren and Volt Nuts: for the first: from the experiences of Warren my project seems to be feasible. Im working now since 2008 on my "project". Eliminating step by step the drawbacks of stability. And learning a lot of real and not ideal parts. I changed from SMD references with large hysteresis and bad stability to plastic DIP devices being about a factor 3 better than SMD. Finally finding out that some buried zener devices have relative low hysteresis. But air humidity was a large issue preventing stability below 10ppm over a year. So finally I try to use devices in hermetically tight cases which eliminates humidity if carefully decoupled from PCB stress. So what is feasible with selected parts and a 3rd order temperature compensation is shown in the attached picture. (ADC13_TC) A AD586LJ device as reference within ADC#13 is measuring a heated LM399 reference via a 2:1 capacitive divider. X-Axis is temperature normalized to 25 degrees celsius -10/+15 degrees measured directly at the AD586 reference. Y-Axis on the left is LM399 measurement value in mV (divided by 2). The red line is the measurement value of the ADC. The blue line the resulting correction with a 3rd order polynominal. Y-Axis on the right side is together with the green line the resulting deviation in uV after calculating out the tempco of the reference. Deviation is about +/-2 uV referenced to 3.4V which gives below +/-1ppm resulting tempco after correction. Unfortunately the slope of the uncompensated TC is relative large (about 1.1ppm). Together with my temperature resolution of about 0.1K/step this will give a relative large temperature step noise of up to 0.6uV/step in the 5V-Range of the ADC. But anyway ADC #13 was the first ADC which I can use for ageing measurements. After a run in phase of nearly 1 year the ageing of ADC #13 stabilized. Currently I compare ADC13 nearly every day with 3 heated references (1 LM399 = LM_2 and 2 LTZ1000A = LTZ_1/2). The last half year the ageing is about 0.5 to 1.5 ppm for 6 months compared to the heated references. See picture ADC13_longterm: X-Axis is day Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1, blue = LTZ1000A #2 Y-axis right is temperature in degree celsius of the temperature sensor near ADC13 reference. By the way: up to now I could not measure any effect which is related to thermocouples. Ok my temperature step noise is still too high. And probably I am using the wrong connectors in my tests: cheap D-Sub connectors where a metal shield is equalizing the temperature of 2 relative close neighboured contacts. With best regards Andreas ----- Original Message ----- From: "WarrenS" <warrensjmail-one@yahoo.com> To: <volt-nuts@febo.com> Sent: Sunday, January 27, 2013 11:23 PM Subject: [volt-nuts] Some questions to zeners (1N823-1N829) > Andreas > > 1st my two cents worth concerning comments of others that are being more > practical than nutty. > > "it is not easy to achieve these specs", > Too True, > If you do not already know this, then I'd suggest you back off and do some > major readjustment of your requirements. > Assuming you already know it is not easy and why, then > IMHO, what you want to make is doable with the right compromises, some of > which being > A little extra power. > a lot of extra time > a lot of test selecting > some extra circuits or S/W > > "your expectations are not realistic " > True, if you think it can be done from a data sheet and a few of off the > shelf parts in a production device. > Not true, if you want to go totally nuts and have the time to do it. > > > "The thermoelectric effects causing error would swamp the performance" > Could be true for some. If you do not already know enough to keep these > effects well under <<0.1PPM then should find another starter project. > > > "To keep everything below the 1 ppm/deg C range you would have to put the > entire circuit in controlled temperature" > Too true, But for these requirements, fortunately you do not have to keep > *Everything* under 1 ppm/deg C > ALL you have to do is to keep the sum total of Everything under 1ppm. Big > difference! > And that is very easy to do, by a fact of ten plus if desired. > > > "long term drift and noise will be intrinsic to the devices, and > unpredictable except in a statistical sense." > Half true, > Using the *right device*, and a lot of time, noise can be low enough to be > mostly insignificant > and long term drift will follow a predicable slope. > It ain't easy and it's going to take a lot of time. If you think > otherwise, you need a new project.. > > > "provide near zero tempco at one temperature only" > Answer too limited to be useful > zero TCing a circuit, can provide no voltage difference at any two > temperatures, (to the limit of the repeatability) > Depending how flat you need it between these two temperatures, is the only > issue. > Best to plan on having a second order TC method as well. > A total max total deviation of 1ppm and a 0.1 PPM /deg is not too hard to > get. > > > "One of the best voltage standard Datron 4910AV (4x LTZ1000) have only 1 > ppm/year drift" > OK, so that shows there is at least one way to do it, so it is not > impossible. > Now just need to find the best way for you to do it. > > "Don't bother with TC zeners" > Not a bad idea, but Unfortunately Not a lot of other choices considering > your requirements. > The other choice you have (that you should consider if time is a big > issue) > is to get some three terminal Fluke voltage references. > If necessary by removing them from old test equipment such as Fluke 731 > Those have already been selected and aged, and you'd just then need to > work on the long term drift compensation selection method. > > > "There are lots of nice IC references available" > True, and if you can live with data sheet specs, they are much better > choices. > But hard if not impossible to find anything that will compare to 1/F pop > corn noise, and long term stability > > > "I doubt that any TCZ will match an LM399" > True when considering a wider temperature range, and it sure makes things > easier, > so a good suggestion if TC was your main problem. > But TC need not be a problem, and most any good zener will outperform most > any selected LM399 in low freq noise and stability. > > > "so would have to be ovenized to get *best* performance". > True of Any circuit when "best" TC is concerned, so not very relevant, > The question is can it be made good enough without an oven? > And the answer is defiantly yes with the right tempco circuit. > And if you want it even better, can make a very low power mini-oven. > > Back to your email ****************** > >> Do you have typical values over a 64-90 ?F range. Will it be above 1ppm/K >> or below? > I do not have any data over that Temp range in front of me. > Over My room changes (about 1/2 that temp range), can keep the Total > voltage error down to around 1PPM, (not just /deg) > If it is an important consideration, I may try and see what happens over a > wider range. > For a backup plan, considering having an addition second order TC > compensating method such as S/W. > Overall compensating to 1ppm / C is no problem, > which is Total change +- 7PPM over a +- 7 deg C range > Ten times better is possible over narrow temp ranges like that. > > It is the hysteresis and stability that is going to be a limiting factor, > and last I looked, > The zeners I tested had no measurable Hysteresis over a much wider temp > range than that. > >>>From your plot it would be 0.33ppm per 3 degrees in narrow range i.e. 0.1 >>>ppm per degree. > Correct, but not a relevant measurement. > ANY straight line TC as seen there can be zeroed out. > That was just a pre-test plot to see if the part was low enough noise to > do a more accurate Zero TC. > >> By the way: is it degrees Fahrenheit or degrees Celsius (= 3 Kelvin)? > My world is mostly in F, and since I was not even plotting temp at the > same time, pretty irrelevant for that data set. > >> If I understand you right then you would not use this device because it >> does >> behave other than the others? > Your understaning is not correct, not even close. My comment applies to a > preselection process of 1N82x parts. > In the case you stated, your preselection process (assuming you have > enough to pick from) would be to select all the parts that are less than > say 1PPM, with maybe a 20% yield. > > If the yield is too low then make a second pass for parts that are say 2 > +-1 PPM, > and in that case, then yes the "best" part would not go into that batch, > because it will likely need a different compensation than the others. > > >> On the other side it seems to be the device with the largest ageing rate >> of the 5 pieces. > likely too little data to be important, Just as likely random luck as > anything important. > Until after you do some pre-aging, (whatever that may mean for that part) > I would not even bother looking at that data this early on in the > selection process. > > >> So I still hope that anyone has experiences with hysteresis of the >> zeners. > Turn on after power down, and hysteresis, repeatable, etc altogether is > under 1PPM. > As part of my pre-test TC procedure, I hit them with cold spray to 0 C and > heat them with a heat gun to ~50C, a few times, > If they are not repeatable they are not used in further testing. > > I have a question about two of your requirements > AJ> tempco below 1ppm/K > AJ> hysteresis in the 10-40 degree range well below 1ppm > > This suggest to me that what you are really planning on making is > something with an overall compensted TC that is well below a 1 ppm/K, > otherwise your hysteresis requirement does not make a lot of sense to me. > So if what you are really after is more like < 1 PPM total error over time > and temp, all this extra trouble now makes a lot more sense. > > > Now the disclaimer, I have no idea if any of the now available 1N825 or > better parts work like I've described. > It is VERY much a manufacture sensitive thing. > Short term noise being the biggest rejecting thing I've seen in the past. > 0.5 PPM (3uv steps) are not untypical for some batches. > > If someone that has enough of these parts from a single batch and > manufacture to make it worth while testing them, and will make them > available to others, then I'll test and report the results and compare > them to what I have. > > There may be another less direct way to get where you want to go (You did > not say where that was) > Divide the task into two or more sections. > As an example, One being the low power short term stable DVM device that > works over a limited temperature. > There would be no problem getting 0.1 PPM accuracy/repeatable for that > part. > and a second higher power device that is mostly off, and powered up Just > long enough to calibrate the low power xfer DVM above. > > And if you can then include some sort of cal exchange program for the > higher powered but less used device, this would make a very hard project > into a relative easy project. > > > ws > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
Михаил
Tue, Jan 29, 2013 7:31 AM

Hi, Andreas!

AJ> But anyway ADC #13 was the first ADC which I can use for ageing
AJ> measurements.

Where to find information about this ADC? It is very interesting thing. Is it
Multi-slope III, PWM, or another type?

Regards,
Mickle T.

Hi, Andreas! AJ> But anyway ADC #13 was the first ADC which I can use for ageing AJ> measurements. Where to find information about this ADC? It is very interesting thing. Is it Multi-slope III, PWM, or another type? Regards, Mickle T.
PK
Poul-Henning Kamp
Tue, Jan 29, 2013 7:38 AM

Content-Type: text/plain; charset=ISO-8859-1

In message AA473A4936944A1199FC6F4C9C5D9A1A@laptop, "Andreas Jahn" writes:

Amazing!

Recently I came across a very interesting ADC from TI, designed for
use with geophones: ADS1282

It might be interesting for volt-nuttery:  It samples 2 bits at
4MHz, which are then downsampled to 31 bits at 4kHz or less.  For
this reason it is incredibly linear and has very good S/N.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

Content-Type: text/plain; charset=ISO-8859-1 -------- In message <AA473A4936944A1199FC6F4C9C5D9A1A@laptop>, "Andreas Jahn" writes: Amazing! Recently I came across a very interesting ADC from TI, designed for use with geophones: ADS1282 It might be interesting for volt-nuttery: It samples 2 bits at 4MHz, which are then downsampled to 31 bits at 4kHz or less. For this reason it is incredibly linear and has very good S/N. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
Тимофеев Михаил
Tue, Jan 29, 2013 9:43 AM

The ADS1282 offers outstanding specification, but not very useful in the wide-scale DVM application range due to low input and reference voltages.
Within a year I'm trying to develop a design of the selfcal 7.5-digits DMM with ADS1282 ADC. Reference (+/- 10V) is a aged 1N829A with 0.03ppm noise and 2nd order TC correction, downscaled to +/-2.5 V with zero TC divider.

Mickle T.

29.01.2013, 11:38, "Poul-Henning Kamp" phk@phk.freebsd.dk:

Content-Type: text/plain; charset=ISO-8859-1

In message AA473A4936944A1199FC6F4C9C5D9A1A@laptop, "Andreas Jahn" writes:

Amazing!

Recently I came across a very interesting ADC from TI, designed for
use with geophones: ADS1282

It might be interesting for volt-nuttery:  It samples 2 bits at
4MHz, which are then downsampled to 31 bits at 4kHz or less.  For
this reason it is incredibly linear and has very good S/N.

--
Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG         | TCP/IP since RFC 956
FreeBSD committer       | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The ADS1282 offers outstanding specification, but not very useful in the wide-scale DVM application range due to low input and reference voltages. Within a year I'm trying to develop a design of the selfcal 7.5-digits DMM with ADS1282 ADC. Reference (+/- 10V) is a aged 1N829A with 0.03ppm noise and 2nd order TC correction, downscaled to +/-2.5 V with zero TC divider. Mickle T. 29.01.2013, 11:38, "Poul-Henning Kamp" <phk@phk.freebsd.dk>: > Content-Type: text/plain; charset=ISO-8859-1 > -------- > In message <AA473A4936944A1199FC6F4C9C5D9A1A@laptop>, "Andreas Jahn" writes: > > Amazing! > > Recently I came across a very interesting ADC from TI, designed for > use with geophones: ADS1282 > > It might be interesting for volt-nuttery:  It samples 2 bits at > 4MHz, which are then downsampled to 31 bits at 4kHz or less.  For > this reason it is incredibly linear and has very good S/N. > > -- > Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG         | TCP/IP since RFC 956 > FreeBSD committer       | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
AB
Andy Bardagjy
Tue, Jan 29, 2013 5:31 PM

The ADS1282 is a really cool part! After studying the ADS1282
datasheet is appears that at best it can achieve 130dB SNR which, by
my calculations is effectively 21.3 bits - by the formula
SNR=(6.02N+1.786)dB.

This got me thinking, are there any higher performing "24 bit" ADCs we
should look at for our "nuttery". One that I found is the Linear
LTC2440 (and others in the family). It claims, at its highest
resolution, 24.6 effective bits. That said, I am having trouble
comparing SNR directly. I need to read more to understand how to
convert its noise spec - 200nV RMS noise - to a SNR dB figure.

Andy Bardagjy
bardagjy.com

On Tue, Jan 29, 2013 at 4:43 AM, Тимофеев Михаил timka2k@yandex.ru wrote:

The ADS1282 offers outstanding specification, but not very useful in the wide-scale DVM application range due to low input and reference voltages.
Within a year I'm trying to develop a design of the selfcal 7.5-digits DMM with ADS1282 ADC. Reference (+/- 10V) is a aged 1N829A with 0.03ppm noise and 2nd order TC correction, downscaled to +/-2.5 V with zero TC divider.

Mickle T.

29.01.2013, 11:38, "Poul-Henning Kamp" phk@phk.freebsd.dk:

Content-Type: text/plain; charset=ISO-8859-1

In message AA473A4936944A1199FC6F4C9C5D9A1A@laptop, "Andreas Jahn" writes:

Amazing!

Recently I came across a very interesting ADC from TI, designed for
use with geophones: ADS1282

It might be interesting for volt-nuttery:  It samples 2 bits at
4MHz, which are then downsampled to 31 bits at 4kHz or less.  For
this reason it is incredibly linear and has very good S/N.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The ADS1282 is a really cool part! After studying the ADS1282 datasheet is appears that *at best* it can achieve 130dB SNR which, by my calculations is effectively 21.3 bits - by the formula SNR=(6.02N+1.786)dB. This got me thinking, are there any higher performing "24 bit" ADCs we should look at for our "nuttery". One that I found is the Linear LTC2440 (and others in the family). It claims, at its highest resolution, 24.6 effective bits. That said, I am having trouble comparing SNR directly. I need to read more to understand how to convert its noise spec - 200nV RMS noise - to a SNR dB figure. Andy Bardagjy bardagjy.com On Tue, Jan 29, 2013 at 4:43 AM, Тимофеев Михаил <timka2k@yandex.ru> wrote: > The ADS1282 offers outstanding specification, but not very useful in the wide-scale DVM application range due to low input and reference voltages. > Within a year I'm trying to develop a design of the selfcal 7.5-digits DMM with ADS1282 ADC. Reference (+/- 10V) is a aged 1N829A with 0.03ppm noise and 2nd order TC correction, downscaled to +/-2.5 V with zero TC divider. > > Mickle T. > > > 29.01.2013, 11:38, "Poul-Henning Kamp" <phk@phk.freebsd.dk>: >> Content-Type: text/plain; charset=ISO-8859-1 >> -------- >> In message <AA473A4936944A1199FC6F4C9C5D9A1A@laptop>, "Andreas Jahn" writes: >> >> Amazing! >> >> Recently I came across a very interesting ADC from TI, designed for >> use with geophones: ADS1282 >> >> It might be interesting for volt-nuttery: It samples 2 bits at >> 4MHz, which are then downsampled to 31 bits at 4kHz or less. For >> this reason it is incredibly linear and has very good S/N. >> >> -- >> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 >> phk@FreeBSD.ORG | TCP/IP since RFC 956 >> FreeBSD committer | BSD since 4.3-tahoe >> Never attribute to malice what can adequately be explained by incompetence. >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
B
beale
Tue, Jan 29, 2013 6:49 PM

-------Original Message-------
From: Poul-Henning Kamp phk@phk.freebsd.dk
...a very interesting ADC from TI, designed for use with geophones: ADS1282

31 bits is impressive. With the +/-2.5 V reference I gather the maximum input range in differential mode is Vref/2 or +/- 1.25 V. The Voffset drift is 0.02 uV/C "typical" and the INL is 0.5 ppm "typical", but 4 ppm "max".  For comparison, the 24-bit Analog Devices AD7190 claims typical Voffset = 0.1 uV/C, or 0.005 uV/C in chop mode, and INL = 1 ppm typ. and 5 ppm max, with a maximum differential input voltage of +/-5 V. So the AD7190 has 2x worse INL (typ.), but with a 4x larger input voltage range, the INL is 2x better (typical) and almost 4x better (max) when measured in microvolts.

For quantity 1 purchase, the AD7190 is $11 (Newark/Element14) and the ADS1282 is $62 (Arrow). For $62, you can get a complete eval board for the AD7190 which works as a 4-channel input standalone USB-powered voltmeter (although the eval board's ADR421 2.5V reference is not volt-nut quality.)  By the way, the 7190 chip provides for two selectable, independent differential Vref inputs, which are separate from the two differential or four pseudo-differential signal inputs.

> -------Original Message------- > From: Poul-Henning Kamp <phk@phk.freebsd.dk> > ...a very interesting ADC from TI, designed for use with geophones: ADS1282 31 bits is impressive. With the +/-2.5 V reference I gather the maximum input range in differential mode is Vref/2 or +/- 1.25 V. The Voffset drift is 0.02 uV/C "typical" and the INL is 0.5 ppm "typical", but 4 ppm "max". For comparison, the 24-bit Analog Devices AD7190 claims typical Voffset = 0.1 uV/C, or 0.005 uV/C in chop mode, and INL = 1 ppm typ. and 5 ppm max, with a maximum differential input voltage of +/-5 V. So the AD7190 has 2x worse INL (typ.), but with a 4x larger input voltage range, the INL is 2x better (typical) and almost 4x better (max) when measured in microvolts. For quantity 1 purchase, the AD7190 is $11 (Newark/Element14) and the ADS1282 is $62 (Arrow). For $62, you can get a complete eval board for the AD7190 which works as a 4-channel input standalone USB-powered voltmeter (although the eval board's ADR421 2.5V reference is not volt-nut quality.) By the way, the 7190 chip provides for two selectable, independent differential Vref inputs, which are separate from the two differential or four pseudo-differential signal inputs.
TH
Tony Holt
Wed, Jan 30, 2013 9:49 AM

On 28/01/2013 22:47, Andreas Jahn wrote:

After a run in phase of nearly 1 year the ageing of ADC #13 stabilized.
Currently I compare ADC13 nearly every day with 3 heated references (1
LM399 = LM_2 and 2 LTZ1000A = LTZ_1/2).
The last half year the ageing is about 0.5 to 1.5 ppm for 6 months
compared to the heated references.
See picture ADC13_longterm:
X-Axis is day
Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1,
blue = LTZ1000A #2
Y-axis right is temperature in degree celsius of the temperature
sensor near ADC13 reference.

By the way: up to now I could not measure any effect which is related
to thermocouples.
Ok my temperature step noise is still too high. And probably I am
using the wrong connectors in my tests:
cheap D-Sub connectors where a metal shield is equalizing the
temperature of 2 relative close neighboured contacts.

With best regards

Andreas

Andreas,

Very interesting results - thanks for sharing your painstaking work.
Hope you don't mind me asking a few questions though:

How are you dealing with the issue of drift in the thermocouple
measurements (including the cold junction compensation)? Do you
calibrate it periodically? Thermocouples aren't noted for high stability

  • but presumably at room temperature its perhaps not much of an issue.

Do you know what temperature the LTZ1000 references are operating at,
and how long have they been operated for - ie. have they been aged prior
to starting the long term test? (Presumably the answer to that is the
fact that you are showing results from day 460 onwards?)

Have you any insight into how stable the ADC's reference (AD586LJ) is?
I.E. Have you made any occasional or periodic measurements with other
calibrated instruments during the long term test or is it the long term
test results themselves which leads you to state: "After a run in phase
of nearly 1 year the ageing of ADC #13 stabilized."?

Thanks, Tony H

On 28/01/2013 22:47, Andreas Jahn wrote: > After a run in phase of nearly 1 year the ageing of ADC #13 stabilized. > Currently I compare ADC13 nearly every day with 3 heated references (1 > LM399 = LM_2 and 2 LTZ1000A = LTZ_1/2). > The last half year the ageing is about 0.5 to 1.5 ppm for 6 months > compared to the heated references. > See picture ADC13_longterm: > X-Axis is day > Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1, > blue = LTZ1000A #2 > Y-axis right is temperature in degree celsius of the temperature > sensor near ADC13 reference. > > By the way: up to now I could not measure any effect which is related > to thermocouples. > Ok my temperature step noise is still too high. And probably I am > using the wrong connectors in my tests: > cheap D-Sub connectors where a metal shield is equalizing the > temperature of 2 relative close neighboured contacts. > > With best regards > > Andreas Andreas, Very interesting results - thanks for sharing your painstaking work. Hope you don't mind me asking a few questions though: How are you dealing with the issue of drift in the thermocouple measurements (including the cold junction compensation)? Do you calibrate it periodically? Thermocouples aren't noted for high stability - but presumably at room temperature its perhaps not much of an issue. Do you know what temperature the LTZ1000 references are operating at, and how long have they been operated for - ie. have they been aged prior to starting the long term test? (Presumably the answer to that is the fact that you are showing results from day 460 onwards?) Have you any insight into how stable the ADC's reference (AD586LJ) is? I.E. Have you made any occasional or periodic measurements with other calibrated instruments during the long term test or is it the long term test results themselves which leads you to state: "After a run in phase of nearly 1 year the ageing of ADC #13 stabilized."? Thanks, Tony H
AJ
Andreas Jahn
Fri, Feb 1, 2013 11:00 PM

Hello Tony,

I do not understand your question with thermocouples.
The temperature measurement within ADC13 is done by a NTC.
The thermocouple discussion came up because of the widely used banana plugs
on instruments.

Most of your questions to LTZ1000 have been answered mid to late 2010 within
this board.
Day zero of this diagram is the day when I started automatic measurement
sequences with a relay multiplexer.
ADC13 was included around day 200. Some weeks after first powering and after
doing all adjustments.
The whole story of ADC13 can be seen in the diagram.
First ageing around 3.4 ppm/khr (light blue) referenced to first powering.
Second ageing 2.4 ppm/khr (pink) due to loading the reference by 15mA over
night.
After some weeks ageing slowed down suddenly. So I terminated the 15mA
loading.

My world are the 3 references 1 LM 399 and 2 LTZ1000A.
I compare them annually to 2 Keithley 2000.
Drift is within 1 count (10uV) over 1 to 2 years. (same direction on both
Keithleys)
LM399 seems to drift upwards. The 2 LTZ1000 seem to drift slightly
downwards.

Out of 29 AD586LQ I found about 3-4 which have usable parameters.

With best regards

Andreas

----- Original Message -----
From: "Tony Holt" vnuts@toneh.demon.co.uk
To: volt-nuts@febo.com
Sent: Wednesday, January 30, 2013 10:49 AM
Subject: Re: [volt-nuts] Some questions to zeners (1N823-1N829)

On 28/01/2013 22:47, Andreas Jahn wrote:

After a run in phase of nearly 1 year the ageing of ADC #13 stabilized.
Currently I compare ADC13 nearly every day with 3 heated references (1
LM399 = LM_2 and 2 LTZ1000A = LTZ_1/2).
The last half year the ageing is about 0.5 to 1.5 ppm for 6 months
compared to the heated references.
See picture ADC13_longterm:
X-Axis is day
Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1, blue
= LTZ1000A #2
Y-axis right is temperature in degree celsius of the temperature sensor
near ADC13 reference.

By the way: up to now I could not measure any effect which is related to
thermocouples.
Ok my temperature step noise is still too high. And probably I am using
the wrong connectors in my tests:
cheap D-Sub connectors where a metal shield is equalizing the temperature
of 2 relative close neighboured contacts.

With best regards

Andreas

Andreas,

Very interesting results - thanks for sharing your painstaking work. Hope
you don't mind me asking a few questions though:

How are you dealing with the issue of drift in the thermocouple
measurements (including the cold junction compensation)? Do you calibrate
it periodically? Thermocouples aren't noted for high stability - but
presumably at room temperature its perhaps not much of an issue.

Do you know what temperature the LTZ1000 references are operating at, and
how long have they been operated for - ie. have they been aged prior to
starting the long term test? (Presumably the answer to that is the fact
that you are showing results from day 460 onwards?)

Have you any insight into how stable the ADC's reference (AD586LJ) is?
I.E. Have you made any occasional or periodic measurements with other
calibrated instruments during the long term test or is it the long term
test results themselves which leads you to state: "After a run in phase of
nearly 1 year the ageing of ADC #13 stabilized."?

Thanks, Tony H


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hello Tony, I do not understand your question with thermocouples. The temperature measurement within ADC13 is done by a NTC. The thermocouple discussion came up because of the widely used banana plugs on instruments. Most of your questions to LTZ1000 have been answered mid to late 2010 within this board. Day zero of this diagram is the day when I started automatic measurement sequences with a relay multiplexer. ADC13 was included around day 200. Some weeks after first powering and after doing all adjustments. The whole story of ADC13 can be seen in the diagram. First ageing around 3.4 ppm/khr (light blue) referenced to first powering. Second ageing 2.4 ppm/khr (pink) due to loading the reference by 15mA over night. After some weeks ageing slowed down suddenly. So I terminated the 15mA loading. My world are the 3 references 1 LM 399 and 2 LTZ1000A. I compare them annually to 2 Keithley 2000. Drift is within 1 count (10uV) over 1 to 2 years. (same direction on both Keithleys) LM399 seems to drift upwards. The 2 LTZ1000 seem to drift slightly downwards. Out of 29 AD586LQ I found about 3-4 which have usable parameters. With best regards Andreas ----- Original Message ----- From: "Tony Holt" <vnuts@toneh.demon.co.uk> To: <volt-nuts@febo.com> Sent: Wednesday, January 30, 2013 10:49 AM Subject: Re: [volt-nuts] Some questions to zeners (1N823-1N829) > On 28/01/2013 22:47, Andreas Jahn wrote: >> After a run in phase of nearly 1 year the ageing of ADC #13 stabilized. >> Currently I compare ADC13 nearly every day with 3 heated references (1 >> LM399 = LM_2 and 2 LTZ1000A = LTZ_1/2). >> The last half year the ageing is about 0.5 to 1.5 ppm for 6 months >> compared to the heated references. >> See picture ADC13_longterm: >> X-Axis is day >> Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1, blue >> = LTZ1000A #2 >> Y-axis right is temperature in degree celsius of the temperature sensor >> near ADC13 reference. >> >> By the way: up to now I could not measure any effect which is related to >> thermocouples. >> Ok my temperature step noise is still too high. And probably I am using >> the wrong connectors in my tests: >> cheap D-Sub connectors where a metal shield is equalizing the temperature >> of 2 relative close neighboured contacts. >> >> With best regards >> >> Andreas > Andreas, > > Very interesting results - thanks for sharing your painstaking work. Hope > you don't mind me asking a few questions though: > > How are you dealing with the issue of drift in the thermocouple > measurements (including the cold junction compensation)? Do you calibrate > it periodically? Thermocouples aren't noted for high stability - but > presumably at room temperature its perhaps not much of an issue. > > Do you know what temperature the LTZ1000 references are operating at, and > how long have they been operated for - ie. have they been aged prior to > starting the long term test? (Presumably the answer to that is the fact > that you are showing results from day 460 onwards?) > > Have you any insight into how stable the ADC's reference (AD586LJ) is? > I.E. Have you made any occasional or periodic measurements with other > calibrated instruments during the long term test or is it the long term > test results themselves which leads you to state: "After a run in phase of > nearly 1 year the ageing of ADC #13 stabilized."? > > Thanks, Tony H > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.