time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Stanford Research SR620 Measurement Bias

S
Stan
Sun, Sep 27, 2009 4:15 PM

I just acquired an SR620 that looks to be in excellent condition. I've
noticed something in its operation and I'm not sure if it represents a
problem, a calibration issue, or if it's just a "feature."

Yesterday, after letting the counter warm up for an hour, I calibrated the
10 MHz timebase and did a couple of Auto-Cals (just to be sure), but even
after all of that, there's still a bias of somewhere between 40 and 70
counts in the frequency readout, independent of the gate time. For example,
if I measure the 5 MHz output of my 5065A, and using a sample size of 1,
here's what I see (these are representative readings-the values are bouncing
up and down by 10 or 20 counts and I'm "visually averaging" them).
Increasing the sample size has no effect other than to reduce the
variability of the measurements, but the bias is still the same magnitude:

1 sec Gate:    5,000,000.00054 Hz

0.1 sec Gate:  5,000,000.0060 Hz

0.01 sec Gate:  5,000,000.055 Hz

1 Period Gate:  5,004,7__ Hz

I've had several HP 5370A/B counters, but never one of these. As I
recall-and it's been a while-there's an adjustment in the 5370A/B for the
interpolators that can remove this sort of bias. Is there a similar
adjustment in the SR620? I've been through the manual and don't see anything
obvious.

Thanks for any advice!

Stan

I just acquired an SR620 that looks to be in excellent condition. I've noticed something in its operation and I'm not sure if it represents a problem, a calibration issue, or if it's just a "feature." Yesterday, after letting the counter warm up for an hour, I calibrated the 10 MHz timebase and did a couple of Auto-Cals (just to be sure), but even after all of that, there's still a bias of somewhere between 40 and 70 counts in the frequency readout, independent of the gate time. For example, if I measure the 5 MHz output of my 5065A, and using a sample size of 1, here's what I see (these are representative readings-the values are bouncing up and down by 10 or 20 counts and I'm "visually averaging" them). Increasing the sample size has no effect other than to reduce the variability of the measurements, but the bias is still the same magnitude: 1 sec Gate: 5,000,000.00054 Hz 0.1 sec Gate: 5,000,000.0060 Hz 0.01 sec Gate: 5,000,000.055 Hz 1 Period Gate: 5,004,7__ Hz I've had several HP 5370A/B counters, but never one of these. As I recall-and it's been a while-there's an adjustment in the 5370A/B for the interpolators that can remove this sort of bias. Is there a similar adjustment in the SR620? I've been through the manual and don't see anything obvious. Thanks for any advice! Stan
MD
Magnus Danielson
Sun, Sep 27, 2009 9:55 PM

Stan,

Stan wrote:

I just acquired an SR620 that looks to be in excellent condition. I've
noticed something in its operation and I'm not sure if it represents a
problem, a calibration issue, or if it's just a "feature."

Yesterday, after letting the counter warm up for an hour, I calibrated the
10 MHz timebase and did a couple of Auto-Cals (just to be sure), but even
after all of that, there's still a bias of somewhere between 40 and 70
counts in the frequency readout, independent of the gate time. For example,
if I measure the 5 MHz output of my 5065A, and using a sample size of 1,
here's what I see (these are representative readings-the values are bouncing
up and down by 10 or 20 counts and I'm "visually averaging" them).
Increasing the sample size has no effect other than to reduce the
variability of the measurements, but the bias is still the same magnitude:

1 sec Gate:    5,000,000.00054 Hz

0.1 sec Gate:  5,000,000.0060 Hz

0.01 sec Gate:  5,000,000.055 Hz

1 Period Gate:  5,004,7__ Hz

Notice how the bias shift up one decade as the time-base shifts down a
decade. Consider that this is a reciprocal counter which measures the
number of events and amount of time (in its time-base):

freq = Events / Time

As we step Time a decade down, the error increases by a decade, but an
offset in number of Events should produce a much larger error (5 Hz, 50
Hz and 500 Hz respectively) so we must look at the time side of the
division.

To facilitate this, I make up estimates for the Event counter and Gate
Counter and runs the calculation backwards:

freq = Events / Time = Events / (GateCount * 90 MHz - Start + Stop)

Events = freq * Time

Time = Events / freq

tau    Events    Gate Count  -Start + Stop
1 s  5.000.000 90.000.000  -107,999 ps
100 ms  500.000  9.000.000  -119,999 ps
10 ms    50.000    900.000  -109,999 ps
500 ns        1        18  -199,999 ps

This is fairly consistent. Notice that the 1 cycle path is different
from the internally gated path.

However, the cause of this time error needs some more thought.

If it where an inconsistens between the start and stop interpolators,
the error would shift as the internal time-base beats against the
internal 10 MHz oscillator (which is stepped up to the 90 MHz coarse
counter clock).

However, a voltage error between start and stop trigger voltage would
produce a stable offset. Since the start and stop input selection is the
same in the SR620 when doing frequency measurments, this flaw is
canceled out.

The ST620 uses a little dedicated circuit to post-process the frequency
measurements and produce the start and stop signals being sent to
measurement channel muxes. After this mux signals goes to the event and
gate counter setups and also to the interpolator logics. Any systematic
time-offset due to propagation delay variations in the start and stop
delay of that circuit will introduce a time-bias into frequency
measurements.

Looking at the overview of calibration bytes, byte 50 looks like the
byte of interest. However, it is claimed that byte 50 is among those
that is trimmed by AutoCal, which makes the peculiar statement on page
78 that Frequency does not need calibration. It should say it does not
need manual calibration. It may be that you could tweak this value and
see how it changes your readings.

I've had several HP 5370A/B counters, but never one of these. As I
recall-and it's been a while-there's an adjustment in the 5370A/B for the
interpolators that can remove this sort of bias. Is there a similar
adjustment in the SR620? I've been through the manual and don't see anything
obvious.

It is hidden in the details. I think you with the brief analysis above
should be able to come to the same conclusion.

Thanks for any advice!

Hope I got you onto the track.

Cheers,
Magnus

Stan, Stan wrote: > I just acquired an SR620 that looks to be in excellent condition. I've > noticed something in its operation and I'm not sure if it represents a > problem, a calibration issue, or if it's just a "feature." > > Yesterday, after letting the counter warm up for an hour, I calibrated the > 10 MHz timebase and did a couple of Auto-Cals (just to be sure), but even > after all of that, there's still a bias of somewhere between 40 and 70 > counts in the frequency readout, independent of the gate time. For example, > if I measure the 5 MHz output of my 5065A, and using a sample size of 1, > here's what I see (these are representative readings-the values are bouncing > up and down by 10 or 20 counts and I'm "visually averaging" them). > Increasing the sample size has no effect other than to reduce the > variability of the measurements, but the bias is still the same magnitude: > > > > 1 sec Gate: 5,000,000.00054 Hz > > 0.1 sec Gate: 5,000,000.0060 Hz > > 0.01 sec Gate: 5,000,000.055 Hz > > 1 Period Gate: 5,004,7__ Hz Notice how the bias shift up one decade as the time-base shifts down a decade. Consider that this is a reciprocal counter which measures the number of events and amount of time (in its time-base): freq = Events / Time As we step Time a decade down, the error increases by a decade, but an offset in number of Events should produce a much larger error (5 Hz, 50 Hz and 500 Hz respectively) so we must look at the time side of the division. To facilitate this, I make up estimates for the Event counter and Gate Counter and runs the calculation backwards: freq = Events / Time = Events / (GateCount * 90 MHz - Start + Stop) Events = freq * Time Time = Events / freq tau Events Gate Count -Start + Stop 1 s 5.000.000 90.000.000 -107,999 ps 100 ms 500.000 9.000.000 -119,999 ps 10 ms 50.000 900.000 -109,999 ps 500 ns 1 18 -199,999 ps This is fairly consistent. Notice that the 1 cycle path is different from the internally gated path. However, the cause of this time error needs some more thought. If it where an inconsistens between the start and stop interpolators, the error would shift as the internal time-base beats against the internal 10 MHz oscillator (which is stepped up to the 90 MHz coarse counter clock). However, a voltage error between start and stop trigger voltage would produce a stable offset. Since the start and stop input selection is the same in the SR620 when doing frequency measurments, this flaw is canceled out. The ST620 uses a little dedicated circuit to post-process the frequency measurements and produce the start and stop signals being sent to measurement channel muxes. After this mux signals goes to the event and gate counter setups and also to the interpolator logics. Any systematic time-offset due to propagation delay variations in the start and stop delay of that circuit will introduce a time-bias into frequency measurements. Looking at the overview of calibration bytes, byte 50 looks like the byte of interest. However, it is claimed that byte 50 is among those that is trimmed by AutoCal, which makes the peculiar statement on page 78 that Frequency does not need calibration. It should say it does not need manual calibration. It may be that you could tweak this value and see how it changes your readings. > I've had several HP 5370A/B counters, but never one of these. As I > recall-and it's been a while-there's an adjustment in the 5370A/B for the > interpolators that can remove this sort of bias. Is there a similar > adjustment in the SR620? I've been through the manual and don't see anything > obvious. It is hidden in the details. I think you with the brief analysis above should be able to come to the same conclusion. > Thanks for any advice! Hope I got you onto the track. Cheers, Magnus
UB
Ulrich Bangert
Mon, Sep 28, 2009 7:08 AM

Stan,

I own a SR620 as well and mine behaves the same (while with a different
number of "counts") so your's is surely not defective! My understanding a
few years ago when I acquired the counter was that this bias is not due to a
"adjustment" in the normal sense but to set a parameter in the counter's
eeprom to a correct value. Since this parameter is limited to a certain
precisison you can make some effort in minimizing this bias but you should
not expect to make it exactly zero. The manual explains how to do but you do
not find the procedure where you would have expected it.

Best regards
Ulrich Bangert

-----Ursprungliche Nachricht-----
Von: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] Im Auftrag von Stan
Gesendet: Sonntag, 27. September 2009 18:15
An: time-nuts@febo.com
Betreff: [time-nuts] Stanford Research SR620 Measurement Bias

I just acquired an SR620 that looks to be in excellent
condition. I've noticed something in its operation and I'm
not sure if it represents a problem, a calibration issue, or
if it's just a "feature."

Yesterday, after letting the counter warm up for an hour, I
calibrated the 10 MHz timebase and did a couple of Auto-Cals
(just to be sure), but even after all of that, there's still
a bias of somewhere between 40 and 70 counts in the frequency
readout, independent of the gate time. For example, if I
measure the 5 MHz output of my 5065A, and using a sample size
of 1, here's what I see (these are representative
readings-the values are bouncing up and down by 10 or 20
counts and I'm "visually averaging" them). Increasing the
sample size has no effect other than to reduce the
variability of the measurements, but the bias is still the
same magnitude:

1 sec Gate:    5,000,000.00054 Hz

0.1 sec Gate:  5,000,000.0060 Hz

0.01 sec Gate:  5,000,000.055 Hz

1 Period Gate:  5,004,7__ Hz

I've had several HP 5370A/B counters, but never one of these.
As I recall-and it's been a while-there's an adjustment in
the 5370A/B for the interpolators that can remove this sort
of bias. Is there a similar adjustment in the SR620? I've
been through the manual and don't see anything obvious.

Thanks for any advice!

Stan


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Stan, I own a SR620 as well and mine behaves the same (while with a different number of "counts") so your's is surely not defective! My understanding a few years ago when I acquired the counter was that this bias is not due to a "adjustment" in the normal sense but to set a parameter in the counter's eeprom to a correct value. Since this parameter is limited to a certain precisison you can make some effort in minimizing this bias but you should not expect to make it exactly zero. The manual explains how to do but you do not find the procedure where you would have expected it. Best regards Ulrich Bangert > -----Ursprungliche Nachricht----- > Von: time-nuts-bounces@febo.com > [mailto:time-nuts-bounces@febo.com] Im Auftrag von Stan > Gesendet: Sonntag, 27. September 2009 18:15 > An: time-nuts@febo.com > Betreff: [time-nuts] Stanford Research SR620 Measurement Bias > > > I just acquired an SR620 that looks to be in excellent > condition. I've noticed something in its operation and I'm > not sure if it represents a problem, a calibration issue, or > if it's just a "feature." > > > > Yesterday, after letting the counter warm up for an hour, I > calibrated the 10 MHz timebase and did a couple of Auto-Cals > (just to be sure), but even after all of that, there's still > a bias of somewhere between 40 and 70 counts in the frequency > readout, independent of the gate time. For example, if I > measure the 5 MHz output of my 5065A, and using a sample size > of 1, here's what I see (these are representative > readings-the values are bouncing up and down by 10 or 20 > counts and I'm "visually averaging" them). Increasing the > sample size has no effect other than to reduce the > variability of the measurements, but the bias is still the > same magnitude: > > > > 1 sec Gate: 5,000,000.00054 Hz > > 0.1 sec Gate: 5,000,000.0060 Hz > > 0.01 sec Gate: 5,000,000.055 Hz > > 1 Period Gate: 5,004,7__ Hz > > > > I've had several HP 5370A/B counters, but never one of these. > As I recall-and it's been a while-there's an adjustment in > the 5370A/B for the interpolators that can remove this sort > of bias. Is there a similar adjustment in the SR620? I've > been through the manual and don't see anything obvious. > > > > Thanks for any advice! > > > > Stan > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.