View all threadsHi all,
I see an interesting trend and I'm not sure how to explain it ...
I've done a y-factor measurement, sweeping the frequency, sampling
rate, and rx-gain with an x310 with the SBX-120 daughterboard. The
results seem similar to the published performance specs. The part I'm
not sure quite how to explain is why the NF would vary with the
sampling rate. Doesn't the X310 use a static sample rate and
downsample in the FPGA? Why would this affect the NF? Why does the
effect seem to be more exaggerated at higher gain settings? Why is the
effect on NF very small at higher sampling rates but more pronounced
at lower ones instead of being a linear change?
Dustin
On 03/03/2025 22:23, Dustin Widmann via USRP-users wrote:
Hi all,
I see an interesting trend and I'm not sure how to explain it ...
I've done a y-factor measurement, sweeping the frequency, sampling
rate, and rx-gain with an x310 with the SBX-120 daughterboard. The
results seem similar to the published performance specs. The part I'm
not sure quite how to explain is why the NF would vary with the
sampling rate. Doesn't the X310 use a static sample rate and
downsample in the FPGA? Why would this affect the NF? Why does the
effect seem to be more exaggerated at higher gain settings? Why is the
effect on NF very small at higher sampling rates but more pronounced
at lower ones instead of being a linear change?
Dustin
What noise inputs are you using for the two levels in your Y-factor?
The amount of power represented at each sample-rate is different, and
shooting from the hip here, the amount of energy
represented in the transition regions at different sample rates will
be different. For example, odd/even/factor-of-4
sample-rates have different pass-band shapes.
For an ADC, I believe the noise power is fairly constant. As sample rate
increases, noise power density should decrease and SNR should improve. Thus
a measured improvement in noise figure.
Source:
https://www.analog.com/en/resources/technical-articles/noise-spectral-density.html
x310 ADC (ADS62P48): http://www.ti.com/lit/ds/slas635b/slas635b.pdf
Mark
On Mon, Mar 3, 2025, 7:48 PM Marcus D. Leech patchvonbraun@gmail.com
wrote:
On 03/03/2025 22:23, Dustin Widmann via USRP-users wrote:
Hi all,
I see an interesting trend and I'm not sure how to explain it ...
I've done a y-factor measurement, sweeping the frequency, sampling
rate, and rx-gain with an x310 with the SBX-120 daughterboard. The
results seem similar to the published performance specs. The part I'm
not sure quite how to explain is why the NF would vary with the
sampling rate. Doesn't the X310 use a static sample rate and
downsample in the FPGA? Why would this affect the NF? Why does the
effect seem to be more exaggerated at higher gain settings? Why is the
effect on NF very small at higher sampling rates but more pronounced
at lower ones instead of being a linear change?
Dustin
What noise inputs are you using for the two levels in your Y-factor?
The amount of power represented at each sample-rate is different, and
shooting from the hip here, the amount of energy
represented in the transition regions at different sample rates will
be different. For example, odd/even/factor-of-4
sample-rates have different pass-band shapes.
USRP-users mailing list -- usrp-users@lists.ettus.com
To unsubscribe send an email to usrp-users-leave@lists.ettus.com
On 03/03/2025 22:59, Mark Gannet wrote:
For an ADC, I believe the noise power is fairly constant. As sample
rate increases, noise power density should decrease and SNR should
improve. Thus a measured improvement in noise figure.
Source:
https://www.analog.com/en/resources/technical-articles/noise-spectral-density.html
x310 ADC (ADS62P48): http://www.ti.com/lit/ds/slas635b/slas635b.pdf
Mark
Thanks. Trying to get my head around this a bit. The X310 runs the ADC
at a fixed rate, and host-directed sample rates are
based on the actions of a DDC filter/sample-rate converter.
On Mon, Mar 3, 2025, 7:48 PM Marcus D. Leech patchvonbraun@gmail.com
wrote:
On 03/03/2025 22:23, Dustin Widmann via USRP-users wrote:
Hi all,
I see an interesting trend and I'm not sure how to explain it ...
I've done a y-factor measurement, sweeping the frequency, sampling
rate, and rx-gain with an x310 with the SBX-120 daughterboard. The
results seem similar to the published performance specs. The
part I'm
not sure quite how to explain is why the NF would vary with the
sampling rate. Doesn't the X310 use a static sample rate and
downsample in the FPGA? Why would this affect the NF? Why does the
effect seem to be more exaggerated at higher gain settings? Why
is the
effect on NF very small at higher sampling rates but more pronounced
at lower ones instead of being a linear change?
Dustin
What noise inputs are you using for the two levels in your Y-factor?
The amount of power represented at each sample-rate is different, and
shooting from the hip here, the amount of energy
represented in the transition regions at different sample rates
will
be different. For example, odd/even/factor-of-4
sample-rates have different pass-band shapes.
_______________________________________________
USRP-users mailing list -- usrp-users@lists.ettus.com
To unsubscribe send an email to usrp-users-leave@lists.ettus.com
50-ohm termination for cold
~15.5 dB ENR Noise Source for hot
I've attached the scripts and the data, in case they should somehow
prove userful.
Dustin
On Mon, Mar 3, 2025 at 10:49 PM Marcus D. Leech patchvonbraun@gmail.com wrote:
On 03/03/2025 22:23, Dustin Widmann via USRP-users wrote:
Hi all,
I see an interesting trend and I'm not sure how to explain it ...
I've done a y-factor measurement, sweeping the frequency, sampling
rate, and rx-gain with an x310 with the SBX-120 daughterboard. The
results seem similar to the published performance specs. The part I'm
not sure quite how to explain is why the NF would vary with the
sampling rate. Doesn't the X310 use a static sample rate and
downsample in the FPGA? Why would this affect the NF? Why does the
effect seem to be more exaggerated at higher gain settings? Why is the
effect on NF very small at higher sampling rates but more pronounced
at lower ones instead of being a linear change?
Dustin
What noise inputs are you using for the two levels in your Y-factor?
The amount of power represented at each sample-rate is different, and
shooting from the hip here, the amount of energy
represented in the transition regions at different sample rates will
be different. For example, odd/even/factor-of-4
sample-rates have different pass-band shapes.