Discussion of precise voltage measurement

View all threads

Re: [volt-nuts] Solartron 7061 help request (eeprom?)

JL
J. L. Trantham
Sat, Mar 24, 2012 1:11 AM

Mickle,

Do you have a translation to English?

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of ?????? (Mickle)
Sent: Friday, March 23, 2012 1:59 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions
(washelprequest (eeprom?))

Hi, Fabio!
Zener TC correction in the 7061 and 7071/7081 is very simple. Recently I
wrote a
small article about TC correction methods in precision DMMs and calibrators:
http://www.radiokot.ru/forum/viewtopic.php?f=10&t=52829&start=60 (sorry, but
in
Russian). It is possible to achieve 0.1 ppm/C temperature drift by choosing
the
proper coefficients of the quadratic function. But 7061 maintenance manual -
it's amazing!

Mickle T.

Friday, March 23, 2012, 4:23:57 PM, you wrote:
FE> Finally the manual saga ended :
FE> today the 7061 maintenance manual arrived,
FE> it's photocopied (or laser printed maybe),
FE> very clean and readable.

FE> I had only the time to read superficially,
FE> but I liked the description of temperature
FE> compensaiton for the reference diode,
FE> now I only need a controlled climatic chamber
FE> to play with :)

FE> I received as bonus a 7075 operator manual
FE> original (not photocopied), it will be
FE> nice companion to the 7075 that I received
FE> recently.

FE> Fabio.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Mickle, Do you have a translation to English? Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of ?????? (Mickle) Sent: Friday, March 23, 2012 1:59 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions (washelprequest (eeprom?)) Hi, Fabio! Zener TC correction in the 7061 and 7071/7081 is very simple. Recently I wrote a small article about TC correction methods in precision DMMs and calibrators: http://www.radiokot.ru/forum/viewtopic.php?f=10&t=52829&start=60 (sorry, but in Russian). It is possible to achieve 0.1 ppm/C temperature drift by choosing the proper coefficients of the quadratic function. But 7061 maintenance manual - it's amazing! Mickle T. Friday, March 23, 2012, 4:23:57 PM, you wrote: FE> Finally the manual saga ended : FE> today the 7061 maintenance manual arrived, FE> it's photocopied (or laser printed maybe), FE> very clean and readable. FE> I had only the time to read superficially, FE> but I liked the description of temperature FE> compensaiton for the reference diode, FE> now I only need a controlled climatic chamber FE> to play with :) FE> I received as bonus a 7075 operator manual FE> original (not photocopied), it will be FE> nice companion to the 7075 that I received FE> recently. FE> Fabio. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Михаил
Sat, Mar 24, 2012 4:22 AM

Joe, Fabio,
I'll try to translate the article into English.

P.S. Fabio, Solartron 7075 internals is rather interesting! Thank you!

Saturday, March 24, 2012, 5:11:48 AM, you wrote:
JLT> Mickle,
JLT> Do you have a translation to English?
JLT> Joe

Joe, Fabio, I'll try to translate the article into English. P.S. Fabio, Solartron 7075 internals is rather interesting! Thank you! Saturday, March 24, 2012, 5:11:48 AM, you wrote: JLT> Mickle, JLT> Do you have a translation to English? JLT> Joe
RA
Robert Atkinson
Sat, Mar 24, 2012 2:37 PM

Hi,

I'm late to this thread, but never had much luck with telford. Even before the web, they had lists with no prices and were poor at getting back with prices and availability. Maybe if I was buying big ticket items they may have paid more attention.  What some dealers don't realise is the hobby buyer may also have significant input to bigger company purchases.

Robert G8RPI


From: David C. Partridge david.partridge@perdrix.co.uk
To: 'Discussion of precise voltage measurement' volt-nuts@febo.com
Sent: Thursday, 15 March 2012, 18:12
Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions (washelp request (eeprom?))

Hi Fabio

I always found they were better to talk to on the 'phone than dealing with their emails

Dave
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Fabio Eboli
Sent: 15 March 2012 17:43
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions (washelp request (eeprom?))

Another update, maybe if someone wants to buy from Telford wants to know what happens to me:
Telford answered me 6/mar and was helpful, also we agreed on the price (2 manuals instead of one :), 8/mar I sent the payment trough paypal.
12/mar I sent a mail, asking fot aknowledgment of payment, delivery date etc, no more answers so far :( Now I hope at least I'll receive the manuals soon, I'll let you know.
Fabio.

Fabio Eboli FabioEb@quipo.it ha scritto:

A little update: no more answers to my two emails, seems that Telford
doesn't want my money after all ;)

Fabio.

Fred Schneider pa4tim@gmail.com ha scritto:

I send them several mails but I did not get an answer so if you
manage to buy one, let me know how you did it. ;-)

Fred PA4TIM

Op 3 mrt. 2012 om 19:57 heeft Fabio Eboli FabioEb@quipo.it het
volgende geschreven:

Non need to apologize, thank you for your help.
I asked Telford for a quote early this week, the price is not low,
eventually I'll order from them.

Fabio.

"David C. Partridge" david.partridge@perdrix.co.uk ha scritto:

My apologies that was the User Manual.  The only place I know that
has the Maintenance Manual is Telford Electronics in the UK (they
sell photocopies).

Dave
-----Original Message-----
From: volt-nuts-bounces@febo.com
[mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge
Sent: 03 March 2012 18:22
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions
(washelp request (eeprom?))

http://www.ko4bb.com/manuals/index.php will find you the manual

Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com
[mailto:volt-nuts-bounces@febo.com] On Behalf Of Fabio Eboli
Sent: 03 March 2012 14:46
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions
(was help request (eeprom?))

Little update: tried shorting SP2 jumper but the meter stops
reading DC volts, like if the input is disconnected, no more tests
on this, I disconnected the jumper like as before...
strange ?

Fabio.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


This message was sent using IMP, the Internet Messaging Program.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


This message was sent using IMP, the Internet Messaging Program.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


This message was sent using IMP, the Internet Messaging Program.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi, I'm late to this thread, but never had much luck with telford. Even before the web, they had lists with no prices and were poor at getting back with prices and availability. Maybe if I was buying big ticket items they may have paid more attention.  What some dealers don't realise is the hobby buyer may also have significant input to bigger company purchases. Robert G8RPI ________________________________ From: David C. Partridge <david.partridge@perdrix.co.uk> To: 'Discussion of precise voltage measurement' <volt-nuts@febo.com> Sent: Thursday, 15 March 2012, 18:12 Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions (washelp request (eeprom?)) Hi Fabio I always found they were better to talk to on the 'phone than dealing with their emails Dave -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Fabio Eboli Sent: 15 March 2012 17:43 To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions (washelp request (eeprom?)) Another update, maybe if someone wants to buy from Telford wants to know what happens to me: Telford answered me 6/mar and was helpful, also we agreed on the price (2 manuals instead of one :), 8/mar I sent the payment trough paypal. 12/mar I sent a mail, asking fot aknowledgment of payment, delivery date etc, no more answers so far :( Now I hope at least I'll receive the manuals soon, I'll let you know. Fabio. Fabio Eboli <FabioEb@quipo.it> ha scritto: > A little update: no more answers to my two emails, seems that Telford > doesn't want my money after all ;) > > Fabio. > > Fred Schneider <pa4tim@gmail.com> ha scritto: > >> I send them several mails but I did not get an answer so if you >> manage to buy one, let me know how you did it. ;-) >> >> Fred PA4TIM >> >> Op 3 mrt. 2012 om 19:57 heeft Fabio Eboli <FabioEb@quipo.it> het >> volgende geschreven: >> >>> Non need to apologize, thank you for your help. >>> I asked Telford for a quote early this week, the price is not low, >>> eventually I'll order from them. >>> >>> Fabio. >>> >>> "David C. Partridge" <david.partridge@perdrix.co.uk> ha scritto: >>> >>>> My apologies that was the User Manual.  The only place I know that >>>> has the Maintenance Manual is Telford Electronics in the UK (they >>>> sell photocopies). >>>> >>>> Dave >>>> -----Original Message----- >>>> From: volt-nuts-bounces@febo.com >>>> [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge >>>> Sent: 03 March 2012 18:22 >>>> To: 'Discussion of precise voltage measurement' >>>> Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions >>>> (washelp request (eeprom?)) >>>> >>>> <http://www.ko4bb.com/manuals/index.php> will find you the manual >>>> >>>> >>>> Regards, >>>> David Partridge >>>> -----Original Message----- >>>> From: volt-nuts-bounces@febo.com >>>> [mailto:volt-nuts-bounces@febo.com] On Behalf Of Fabio Eboli >>>> Sent: 03 March 2012 14:46 >>>> To: Discussion of precise voltage measurement >>>> Subject: Re: [volt-nuts] Solartron 7061 ROM and behaviour questions >>>> (was help request (eeprom?)) >>>> >>>> Little update: tried shorting SP2 jumper but the meter stops >>>> reading DC volts, like if the input is disconnected, no more tests >>>> on this, I disconnected the jumper like as before... >>>> strange ? >>>> >>>> Fabio. >>>> >>>> >>>> _______________________________________________ >>>> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>>> and follow the instructions there. >>>> >>>> >>>> _______________________________________________ >>>> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>>> and follow the instructions there. >>>> >>> >>> >>> >>> ---------------------------------------------------------------- >>> This message was sent using IMP, the Internet Messaging Program. >>> >>> >>> >>> _______________________________________________ >>> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > > > > ---------------------------------------------------------------- > This message was sent using IMP, the Internet Messaging Program. > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ---------------------------------------------------------------- This message was sent using IMP, the Internet Messaging Program. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Михаил
Wed, Mar 28, 2012 5:20 PM

Hi, Fabio!
Can you make a scan copy of the Solartron 7061 maintenance manual pages with the
description of temperature compensation for the reference diode? It is very
interesting! Are the Zener temperature tests was described in the manual?

Mickle T.

Friday, March 23, 2012, 4:23:57 PM, you wrote:
FE> Finally the manual saga ended :
FE> today the 7061 maintenance manual arrived,
FE> it's photocopied (or laser printed maybe),
FE> very clean and readable.

FE> I had only the time to read superficially,
FE> but I liked the description of temperature
FE> compensaiton for the reference diode,
FE> now I only need a controlled climatic chamber
FE> to play with :)

FE> I received as bonus a 7075 operator manual
FE> original (not photocopied), it will be
FE> nice companion to the 7075 that I received
FE> recently.

FE> Fabio.

FE> Fabio Eboli FabioEb@quipo.it ha scritto:

Hi, Fabio! Can you make a scan copy of the Solartron 7061 maintenance manual pages with the description of temperature compensation for the reference diode? It is very interesting! Are the Zener temperature tests was described in the manual? Mickle T. Friday, March 23, 2012, 4:23:57 PM, you wrote: FE> Finally the manual saga ended : FE> today the 7061 maintenance manual arrived, FE> it's photocopied (or laser printed maybe), FE> very clean and readable. FE> I had only the time to read superficially, FE> but I liked the description of temperature FE> compensaiton for the reference diode, FE> now I only need a controlled climatic chamber FE> to play with :) FE> I received as bonus a 7075 operator manual FE> original (not photocopied), it will be FE> nice companion to the 7075 that I received FE> recently. FE> Fabio. FE> Fabio Eboli <FabioEb@quipo.it> ha scritto:
FE
Fabio Eboli
Wed, Mar 28, 2012 5:34 PM

Mickle, I didnt have the time to study the manual yet,
I'm working on the 7075 now
(this bastard performs self check correctly and as soon
as it exits from the test it locks :(  ).
I'll scan the relevant pages as soon as I can.

Fabio.

?????? timka2k@yandex.ru ha scritto:

Hi, Fabio!
Can you make a scan copy of the Solartron 7061 maintenance manual
pages with the
description of temperature compensation for the reference diode? It is very
interesting! Are the Zener temperature tests was described in the manual?

Mickle T.

Friday, March 23, 2012, 4:23:57 PM, you wrote:
FE> Finally the manual saga ended :
FE> today the 7061 maintenance manual arrived,
FE> it's photocopied (or laser printed maybe),
FE> very clean and readable.

FE> I had only the time to read superficially,
FE> but I liked the description of temperature
FE> compensaiton for the reference diode,
FE> now I only need a controlled climatic chamber
FE> to play with :)

FE> I received as bonus a 7075 operator manual
FE> original (not photocopied), it will be
FE> nice companion to the 7075 that I received
FE> recently.

FE> Fabio.

FE> Fabio Eboli FabioEb@quipo.it ha scritto:


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


This message was sent using IMP, the Internet Messaging Program.

Mickle, I didnt have the time to study the manual yet, I'm working on the 7075 now (this bastard performs self check correctly and as soon as it exits from the test it locks :( ). I'll scan the relevant pages as soon as I can. Fabio. ?????? <timka2k@yandex.ru> ha scritto: > Hi, Fabio! > Can you make a scan copy of the Solartron 7061 maintenance manual > pages with the > description of temperature compensation for the reference diode? It is very > interesting! Are the Zener temperature tests was described in the manual? > > Mickle T. > > Friday, March 23, 2012, 4:23:57 PM, you wrote: > FE> Finally the manual saga ended : > FE> today the 7061 maintenance manual arrived, > FE> it's photocopied (or laser printed maybe), > FE> very clean and readable. > > FE> I had only the time to read superficially, > FE> but I liked the description of temperature > FE> compensaiton for the reference diode, > FE> now I only need a controlled climatic chamber > FE> to play with :) > > FE> I received as bonus a 7075 operator manual > FE> original (not photocopied), it will be > FE> nice companion to the 7075 that I received > FE> recently. > > FE> Fabio. > > FE> Fabio Eboli <FabioEb@quipo.it> ha scritto: > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ---------------------------------------------------------------- This message was sent using IMP, the Internet Messaging Program.
FE
Fabio Eboli
Thu, Mar 29, 2012 2:37 PM

Mickle, here are 3 pages from the manual.
I'm sorry for the quality.
The pages refer to the eventuality to have to replace
the reference diode, in this case the current has to
be set for the minimum Tc.
So, as far as I understand, there is no "second order"
temperature compensation, only there is hardware selection
of minumum TC diode current.
The reference schematic is a simplified version of 7081's one:
it lacks DAC current selection and TC correction circuitry.
So not that much interesting perhaps :)

http://img692.imageshack.us/img692/9079/7061250.jpg
http://img853.imageshack.us/img853/3325/7061251.jpg
http://img651.imageshack.us/img651/7091/7061fig65.jpg

Remains to be understood what does the instrument every 15minutes
when it makes "drift correction".
Fabio.

?????? timka2k@yandex.ru ha scritto:

Hi, Fabio!
Can you make a scan copy of the Solartron 7061 maintenance manual
pages with the
description of temperature compensation for the reference diode? It is very
interesting! Are the Zener temperature tests was described in the manual?

Mickle T.


This message was sent using IMP, the Internet Messaging Program.

Mickle, here are 3 pages from the manual. I'm sorry for the quality. The pages refer to the eventuality to have to replace the reference diode, in this case the current has to be set for the minimum Tc. So, as far as I understand, there is no "second order" temperature compensation, only there is hardware selection of minumum TC diode current. The reference schematic is a simplified version of 7081's one: it lacks DAC current selection and TC correction circuitry. So not that much interesting perhaps :) http://img692.imageshack.us/img692/9079/7061250.jpg http://img853.imageshack.us/img853/3325/7061251.jpg http://img651.imageshack.us/img651/7091/7061fig65.jpg Remains to be understood what does the instrument every 15minutes when it makes "drift correction". Fabio. ?????? <timka2k@yandex.ru> ha scritto: > Hi, Fabio! > Can you make a scan copy of the Solartron 7061 maintenance manual > pages with the > description of temperature compensation for the reference diode? It is very > interesting! Are the Zener temperature tests was described in the manual? > > Mickle T. > ---------------------------------------------------------------- This message was sent using IMP, the Internet Messaging Program.