PD
Paul DeStefano
Sun, Dec 2, 2012 7:45 PM
All,
The following comment appeared on this list recently and it scared
me a little:
Though the SR620 TIC is a great instrument when hunting the pico seconds
we have to realize, that it's a thermal design desaster (I have to
apologize to all sr620 friends). I have to run it for at least 12
hoursif not 24 to be shure, that every single part is at a more or less
stationary thermal state. Some (NERC) say "...never switch it off".
I assume this instability is due to the instability of the internal
frequency standard. There are two options for the SR620: the standard
oscillator and an ovenized oscillator. In fact, in our measurements, we
plan to use a Cesium frequency standard as the timebase to our SR620.
Does this anecdotal warning apply generally to the instrument or mainly to
the use of the internal standard oscillator?
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.
Many Thanks,
Paul
--
Paul DeStefano
All,
The following comment appeared on this list recently and it scared
me a little:
> Though the SR620 TIC is a great instrument when hunting the pico seconds
> we have to realize, that it's a thermal design desaster (I have to
> apologize to all sr620 friends). I have to run it for at least 12
> hoursif not 24 to be shure, that every single part is at a more or less
> stationary thermal state. Some (NERC) say "...never switch it off".
I assume this instability is due to the instability of the internal
frequency standard. There are two options for the SR620: the standard
oscillator and an ovenized oscillator. In fact, in our measurements, we
plan to use a Cesium frequency standard as the timebase to our SR620.
Does this anecdotal warning apply generally to the instrument or mainly to
the use of the internal standard oscillator?
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.
Many Thanks,
Paul
--
Paul DeStefano
BC
Bob Camp
Sun, Dec 2, 2012 8:12 PM
Hi
Some of the problem comes from the OCXO. Some of it also comes from the stability of the internal circuitry. They do some interesting things with delay lines and the like in the SR620. Best bet is to let it warm up for a while before you need it to perform at it's top level of performance.
Bob
On Dec 2, 2012, at 2:45 PM, Paul DeStefano paul.destefano@willamettealumni.com wrote:
All,
The following comment appeared on this list recently and it scared me a little:
Though the SR620 TIC is a great instrument when hunting the pico seconds we have to realize, that it's a thermal design desaster (I have to apologize to all sr620 friends). I have to run it for at least 12 hoursif not 24 to be shure, that every single part is at a more or less stationary thermal state. Some (NERC) say "...never switch it off".
I assume this instability is due to the instability of the internal frequency standard. There are two options for the SR620: the standard oscillator and an ovenized oscillator. In fact, in our measurements, we plan to use a Cesium frequency standard as the timebase to our SR620. Does this anecdotal warning apply generally to the instrument or mainly to the use of the internal standard oscillator?
We are using the SR620 to measure the interval between 1PPS signals from two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a Rubidium clock.
Many Thanks,
Paul
--
Paul DeStefano
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
Some of the problem comes from the OCXO. Some of it also comes from the stability of the internal circuitry. They do some interesting things with delay lines and the like in the SR620. Best bet is to let it warm up for a while before you need it to perform at it's top level of performance.
Bob
On Dec 2, 2012, at 2:45 PM, Paul DeStefano <paul.destefano@willamettealumni.com> wrote:
> All,
>
> The following comment appeared on this list recently and it scared me a little:
>
>> Though the SR620 TIC is a great instrument when hunting the pico seconds we have to realize, that it's a thermal design desaster (I have to apologize to all sr620 friends). I have to run it for at least 12 hoursif not 24 to be shure, that every single part is at a more or less stationary thermal state. Some (NERC) say "...never switch it off".
>
> I assume this instability is due to the instability of the internal frequency standard. There are two options for the SR620: the standard oscillator and an ovenized oscillator. In fact, in our measurements, we plan to use a Cesium frequency standard as the timebase to our SR620. Does this anecdotal warning apply generally to the instrument or mainly to the use of the internal standard oscillator?
>
> We are using the SR620 to measure the interval between 1PPS signals from two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a Rubidium clock.
>
> Many Thanks,
> Paul
>
> --
> Paul DeStefano
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
CP
Charles P. Steinmetz
Sun, Dec 2, 2012 9:25 PM
The following comment appeared on this list recently and it scared
me a little:
Though the SR620 TIC is a great instrument when hunting the pico
seconds we have to realize, that it's a thermal design desaster (I
have to apologize to all sr620 friends). I have to run it for at
least 12 hoursif not 24 to be shure, that every single part is at a
more or less stationary thermal state. Some (NERC) say "...never
switch it off".
I assume this instability is due to the instability of the internal
frequency standard. * * * In fact, in our measurements, we plan
to use a Cesium frequency standard as the timebase to our SR620.
Does this anecdotal warning apply generally to the instrument or
mainly to the use of the internal standard oscillator?
I concur with the comment above that the thermal design of the 620
could have been better -- the sensing thermistor is in an "exhaust
stack" between the fan (which is blowing out) and the rear enclosure
wall. This means that, instead of trying to maintain the internal
instrument temerature at a constant level, it tries to maintain the
exhaust stack temperature constant with a viciously fast response
time that leads to instability at startup. I have more than once
considered moving the thermistor to a location near the TCXO, but
since the fans always run up to full speed rather quickly at room
temperature anyway, I have never bothered to try to improve the fan circuit.
Additionally. the TCXO remains powered during standby, but not
exactly on frequency because the DAC that adjusts it during operation
is not powered. So, there is some settling from that adding to the
temperature drift. Note also that the DAC steps are not very fine,
so you cannot expect to get the internal oscillator trimmed to better
than e-9 or so. SR apparently thought that most users would connect
620s to external standards, so there was no reason to make them pay
for a high-precision internal standard they would not use.
IME -- operating with an external reference that is better than the
specified accuracy of the 620 -- they meet SR's specifications within
a few minutes at most after switching on from room temperature
storage. (The trigger circuitry may drift a bit as it warms up, so
you may want to check the trigger drift if your application involves
slowish sine waves. I have not investigated this.) Ideally, you
would let the instrument warm up for at least an hour and then
perform an internal calibration before starting your measurements.
All that said, the only way you will know for sure how your
particular instrument and standard will perform is to characterize
them before you start your mobile measurements. In doing so, you
should observe a protocol that resembles the actual travel between
measurements, at least with respect to time and temperature. I
strongly urge you to do this so you can have confidence in your measurements.
Best regards,
Charles
Paul wrote:
>The following comment appeared on this list recently and it scared
>me a little:
>
>>Though the SR620 TIC is a great instrument when hunting the pico
>>seconds we have to realize, that it's a thermal design desaster (I
>>have to apologize to all sr620 friends). I have to run it for at
>>least 12 hoursif not 24 to be shure, that every single part is at a
>>more or less stationary thermal state. Some (NERC) say "...never
>>switch it off".
>
>I assume this instability is due to the instability of the internal
>frequency standard. * * * In fact, in our measurements, we plan
>to use a Cesium frequency standard as the timebase to our SR620.
>Does this anecdotal warning apply generally to the instrument or
>mainly to the use of the internal standard oscillator?
I concur with the comment above that the thermal design of the 620
could have been better -- the sensing thermistor is in an "exhaust
stack" between the fan (which is blowing out) and the rear enclosure
wall. This means that, instead of trying to maintain the internal
instrument temerature at a constant level, it tries to maintain the
exhaust stack temperature constant with a viciously fast response
time that leads to instability at startup. I have more than once
considered moving the thermistor to a location near the TCXO, but
since the fans always run up to full speed rather quickly at room
temperature anyway, I have never bothered to try to improve the fan circuit.
Additionally. the TCXO remains powered during standby, but not
exactly on frequency because the DAC that adjusts it during operation
is not powered. So, there is some settling from that adding to the
temperature drift. Note also that the DAC steps are not very fine,
so you cannot expect to get the internal oscillator trimmed to better
than e-9 or so. SR apparently thought that most users would connect
620s to external standards, so there was no reason to make them pay
for a high-precision internal standard they would not use.
IME -- operating with an external reference that is better than the
specified accuracy of the 620 -- they meet SR's specifications within
a few minutes at most after switching on from room temperature
storage. (The trigger circuitry may drift a bit as it warms up, so
you may want to check the trigger drift if your application involves
slowish sine waves. I have not investigated this.) Ideally, you
would let the instrument warm up for at least an hour and then
perform an internal calibration before starting your measurements.
All that said, the only way you will know for sure how your
particular instrument and standard will perform is to characterize
them before you start your mobile measurements. In doing so, you
should observe a protocol that resembles the actual travel between
measurements, at least with respect to time and temperature. I
strongly urge you to do this so you can have confidence in your measurements.
Best regards,
Charles
MD
Magnus Danielson
Sun, Dec 2, 2012 10:29 PM
On 12/02/2012 10:25 PM, Charles P. Steinmetz wrote:
The following comment appeared on this list recently and it scared me
a little:
Though the SR620 TIC is a great instrument when hunting the pico
seconds we have to realize, that it's a thermal design desaster (I
have to apologize to all sr620 friends). I have to run it for at
least 12 hoursif not 24 to be shure, that every single part is at a
more or less stationary thermal state. Some (NERC) say "...never
switch it off".
I assume this instability is due to the instability of the internal
frequency standard. * * * In fact, in our measurements, we plan to use
a Cesium frequency standard as the timebase to our SR620. Does this
anecdotal warning apply generally to the instrument or mainly to the
use of the internal standard oscillator?
I concur with the comment above that the thermal design of the 620 could
have been better -- the sensing thermistor is in an "exhaust stack"
between the fan (which is blowing out) and the rear enclosure wall. This
means that, instead of trying to maintain the internal instrument
temerature at a constant level, it tries to maintain the exhaust stack
temperature constant with a viciously fast response time that leads to
instability at startup. I have more than once considered moving the
thermistor to a location near the TCXO, but since the fans always run up
to full speed rather quickly at room temperature anyway, I have never
bothered to try to improve the fan circuit.
Never thought about that part, other than the fact that the fan is
annoying like hell.
Additionally. the TCXO remains powered during standby, but not exactly
on frequency because the DAC that adjusts it during operation is not
powered. So, there is some settling from that adding to the temperature
drift. Note also that the DAC steps are not very fine, so you cannot
expect to get the internal oscillator trimmed to better than e-9 or so.
SR apparently thought that most users would connect 620s to external
standards, so there was no reason to make them pay for a high-precision
internal standard they would not use.
Which is why a high stability reference is an option, like most.
In contrast it is interesting to note that the HP5370 had a lower
stability oscillator as option, so removing the 10811 was thus using a
negative option.
IME -- operating with an external reference that is better than the
specified accuracy of the 620 -- they meet SR's specifications within a
few minutes at most after switching on from room temperature storage.
(The trigger circuitry may drift a bit as it warms up, so you may want
to check the trigger drift if your application involves slowish sine
waves. I have not investigated this.) Ideally, you would let the
instrument warm up for at least an hour and then perform an internal
calibration before starting your measurements.
Another source of temperature dependence is the analogue interpolator.
All that said, the only way you will know for sure how your particular
instrument and standard will perform is to characterize them before you
start your mobile measurements. In doing so, you should observe a
protocol that resembles the actual travel between measurements, at least
with respect to time and temperature. I strongly urge you to do this so
you can have confidence in your measurements.
This is good advice. Consider what you need to do. You might want to
consider having a rubidium doing your hold-over. for instance. A PRS-10
would be an interesting option for instance.
Cheers,
Magnus
On 12/02/2012 10:25 PM, Charles P. Steinmetz wrote:
> Paul wrote:
>
>> The following comment appeared on this list recently and it scared me
>> a little:
>>
>>> Though the SR620 TIC is a great instrument when hunting the pico
>>> seconds we have to realize, that it's a thermal design desaster (I
>>> have to apologize to all sr620 friends). I have to run it for at
>>> least 12 hoursif not 24 to be shure, that every single part is at a
>>> more or less stationary thermal state. Some (NERC) say "...never
>>> switch it off".
>>
>> I assume this instability is due to the instability of the internal
>> frequency standard. * * * In fact, in our measurements, we plan to use
>> a Cesium frequency standard as the timebase to our SR620. Does this
>> anecdotal warning apply generally to the instrument or mainly to the
>> use of the internal standard oscillator?
>
> I concur with the comment above that the thermal design of the 620 could
> have been better -- the sensing thermistor is in an "exhaust stack"
> between the fan (which is blowing out) and the rear enclosure wall. This
> means that, instead of trying to maintain the internal instrument
> temerature at a constant level, it tries to maintain the exhaust stack
> temperature constant with a viciously fast response time that leads to
> instability at startup. I have more than once considered moving the
> thermistor to a location near the TCXO, but since the fans always run up
> to full speed rather quickly at room temperature anyway, I have never
> bothered to try to improve the fan circuit.
Never thought about that part, other than the fact that the fan is
annoying like hell.
> Additionally. the TCXO remains powered during standby, but not exactly
> on frequency because the DAC that adjusts it during operation is not
> powered. So, there is some settling from that adding to the temperature
> drift. Note also that the DAC steps are not very fine, so you cannot
> expect to get the internal oscillator trimmed to better than e-9 or so.
> SR apparently thought that most users would connect 620s to external
> standards, so there was no reason to make them pay for a high-precision
> internal standard they would not use.
Which is why a high stability reference is an option, like most.
In contrast it is interesting to note that the HP5370 had a lower
stability oscillator as option, so removing the 10811 was thus using a
negative option.
> IME -- operating with an external reference that is better than the
> specified accuracy of the 620 -- they meet SR's specifications within a
> few minutes at most after switching on from room temperature storage.
> (The trigger circuitry may drift a bit as it warms up, so you may want
> to check the trigger drift if your application involves slowish sine
> waves. I have not investigated this.) Ideally, you would let the
> instrument warm up for at least an hour and then perform an internal
> calibration before starting your measurements.
Another source of temperature dependence is the analogue interpolator.
> All that said, the only way you will know for sure how your particular
> instrument and standard will perform is to characterize them before you
> start your mobile measurements. In doing so, you should observe a
> protocol that resembles the actual travel between measurements, at least
> with respect to time and temperature. I strongly urge you to do this so
> you can have confidence in your measurements.
>
This is good advice. Consider what you need to do. You might want to
consider having a rubidium doing your hold-over. for instance. A PRS-10
would be an interesting option for instance.
Cheers,
Magnus
VE
Volker Esper
Sun, Dec 2, 2012 10:36 PM
Hi,
I was the one who wrote those nasty things about a fantastic instrument.
I bought such a counter a few weeks ago. When I first peeked into it
(because the fan was running at top speed after a few minutes) I
noticed, that the arrangement of the components was anything but optimal.
Examples: The osc is only one or two inches away from the linear voltage
regulators. These are "cooled" by a 1mm iron sheet. The iron sheet has
some cooling slots where the air can flow through, but these slots are
miles away from the regulators. The sheet gets so hot that you won't
touch it voluntarily. If at least the air could flow directly along the
regulators - but they sit in a corner and the air flows diagonally
through the case, avoiding contact with the regulators.
Since I use an external osc, that doesn't bother me too bad.
However, while trying to adjust the trigger circuits, I noticed, that
they are very sensitive to temperature variation. Unfortunately, the
warmed air (which is coming from the cooling slots) flows over these
parts... see picture.
The thermistor that controls the fan is located amidst the air flow -
thus resulting in a very nervous control loop. The fan is continously
speeding up and down. It's not to hard to imagine, that this has effect
on all parts where the air flows (-> Trigger...).
If I had to construct the thermal design for a 4000 Dollar counter I
would do totally different.
That's why I feel a little displeased.
If you own a counter, that you can't modify (due to calibration seals)
heat it up and do not switch it off. Let the cooling air flow by giving
the counter enough space around it. Stabilize room temperature and don't
place your counter near a door or a window or a place where air
temperature is often varying.
I'm shure, you will run a great instrument.
If I own a totally atypical counter and Time Nuts have different
experience please stop me from spreading rubbish.
Volker
Am 02.12.2012 21:12, schrieb Bob Camp:
Hi
Some of the problem comes from the OCXO. Some of it also comes from the stability of the internal circuitry. They do some interesting things with delay lines and the like in the SR620. Best bet is to let it warm up for a while before you need it to perform at it's top level of performance.
Bob
On Dec 2, 2012, at 2:45 PM, Paul DeStefanopaul.destefano@willamettealumni.com wrote:
All,
The following comment appeared on this list recently and it scared me a little:
Though the SR620 TIC is a great instrument when hunting the pico seconds we have to realize, that it's a thermal design desaster (I have to apologize to all sr620 friends). I have to run it for at least 12 hoursif not 24 to be shure, that every single part is at a more or less stationary thermal state. Some (NERC) say "...never switch it off".
I assume this instability is due to the instability of the internal frequency standard. There are two options for the SR620: the standard oscillator and an ovenized oscillator. In fact, in our measurements, we plan to use a Cesium frequency standard as the timebase to our SR620. Does this anecdotal warning apply generally to the instrument or mainly to the use of the internal standard oscillator?
We are using the SR620 to measure the interval between 1PPS signals from two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a Rubidium clock.
Many Thanks,
Paul
--
Paul DeStefano
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi,
I was the one who wrote those nasty things about a fantastic instrument.
I bought such a counter a few weeks ago. When I first peeked into it
(because the fan was running at top speed after a few minutes) I
noticed, that the arrangement of the components was anything but optimal.
Examples: The osc is only one or two inches away from the linear voltage
regulators. These are "cooled" by a 1mm iron sheet. The iron sheet has
some cooling slots where the air can flow through, but these slots are
miles away from the regulators. The sheet gets so hot that you won't
touch it voluntarily. If at least the air could flow directly along the
regulators - but they sit in a corner and the air flows diagonally
through the case, avoiding contact with the regulators.
Since I use an external osc, that doesn't bother me too bad.
However, while trying to adjust the trigger circuits, I noticed, that
they are very sensitive to temperature variation. Unfortunately, the
warmed air (which is coming from the cooling slots) flows over these
parts... see picture.
The thermistor that controls the fan is located amidst the air flow -
thus resulting in a very nervous control loop. The fan is continously
speeding up and down. It's not to hard to imagine, that this has effect
on all parts where the air flows (-> Trigger...).
If I had to construct the thermal design for a 4000 Dollar counter I
would do totally different.
That's why I feel a little displeased.
If you own a counter, that you can't modify (due to calibration seals)
heat it up and do not switch it off. Let the cooling air flow by giving
the counter enough space around it. Stabilize room temperature and don't
place your counter near a door or a window or a place where air
temperature is often varying.
I'm shure, you will run a great instrument.
If I own a totally atypical counter and Time Nuts have different
experience please stop me from spreading rubbish.
Volker
Am 02.12.2012 21:12, schrieb Bob Camp:
> Hi
>
> Some of the problem comes from the OCXO. Some of it also comes from the stability of the internal circuitry. They do some interesting things with delay lines and the like in the SR620. Best bet is to let it warm up for a while before you need it to perform at it's top level of performance.
>
> Bob
>
> On Dec 2, 2012, at 2:45 PM, Paul DeStefano<paul.destefano@willamettealumni.com> wrote:
>
>> All,
>>
>> The following comment appeared on this list recently and it scared me a little:
>>
>>> Though the SR620 TIC is a great instrument when hunting the pico seconds we have to realize, that it's a thermal design desaster (I have to apologize to all sr620 friends). I have to run it for at least 12 hoursif not 24 to be shure, that every single part is at a more or less stationary thermal state. Some (NERC) say "...never switch it off".
>>
>> I assume this instability is due to the instability of the internal frequency standard. There are two options for the SR620: the standard oscillator and an ovenized oscillator. In fact, in our measurements, we plan to use a Cesium frequency standard as the timebase to our SR620. Does this anecdotal warning apply generally to the instrument or mainly to the use of the internal standard oscillator?
>>
>> We are using the SR620 to measure the interval between 1PPS signals from two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a Rubidium clock.
>>
>> Many Thanks,
>> Paul
>>
>> --
>> Paul DeStefano
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
MD
Magnus Danielson
Sun, Dec 2, 2012 11:39 PM
Volker,
On 12/02/2012 11:36 PM, Volker Esper wrote:
Hi,
I was the one who wrote those nasty things about a fantastic instrument.
I bought such a counter a few weeks ago. When I first peeked into it
(because the fan was running at top speed after a few minutes) I
noticed, that the arrangement of the components was anything but optimal.
Examples: The osc is only one or two inches away from the linear voltage
regulators. These are "cooled" by a 1mm iron sheet. The iron sheet has
some cooling slots where the air can flow through, but these slots are
miles away from the regulators. The sheet gets so hot that you won't
touch it voluntarily. If at least the air could flow directly along the
regulators - but they sit in a corner and the air flows diagonally
through the case, avoiding contact with the regulators.
I am also not very impressed with the location of the linear regulators.
This forms a nice interesting relationship between line voltage and room
temperature. As line voltage shifts, the rectified voltage varies, and
then the voltage that the linear regulator needs to "burn off" varies,
and thus their heat contribution varies with the line voltage. Now, the
ambient temperature then controls how easy we get rid of this excess
heat. The way the airflow goes, this cooling makes sure it couples very
well with the electronics, so fan speed amongst other things depend on
ambient temperature... and line voltage. All this assist to give us an
"interesting" temperature dependence.
In contrast, I was impressed by how the HP10811A was located in a
turbolence free corner of the HP5070A/B assembly. While itself also not
particularly well designed in heat context, at least the heat of the
linear regulators was supposed to mostly affect an external
self-convection stream. Forces convection (fan) to stabilize the
temperature of the heat sink would be a good thing there.
Cheers,
Magnus
Volker,
On 12/02/2012 11:36 PM, Volker Esper wrote:
> Hi,
>
> I was the one who wrote those nasty things about a fantastic instrument.
>
> I bought such a counter a few weeks ago. When I first peeked into it
> (because the fan was running at top speed after a few minutes) I
> noticed, that the arrangement of the components was anything but optimal.
>
> Examples: The osc is only one or two inches away from the linear voltage
> regulators. These are "cooled" by a 1mm iron sheet. The iron sheet has
> some cooling slots where the air can flow through, but these slots are
> miles away from the regulators. The sheet gets so hot that you won't
> touch it voluntarily. If at least the air could flow directly along the
> regulators - but they sit in a corner and the air flows diagonally
> through the case, avoiding contact with the regulators.
I am also not very impressed with the location of the linear regulators.
This forms a nice interesting relationship between line voltage and room
temperature. As line voltage shifts, the rectified voltage varies, and
then the voltage that the linear regulator needs to "burn off" varies,
and thus their heat contribution varies with the line voltage. Now, the
ambient temperature then controls how easy we get rid of this excess
heat. The way the airflow goes, this cooling makes sure it couples very
well with the electronics, so fan speed amongst other things depend on
ambient temperature... and line voltage. All this assist to give us an
"interesting" temperature dependence.
In contrast, I was impressed by how the HP10811A was located in a
turbolence free corner of the HP5070A/B assembly. While itself also not
particularly well designed in heat context, at least the heat of the
linear regulators was supposed to mostly affect an external
self-convection stream. Forces convection (fan) to stabilize the
temperature of the heat sink would be a good thing there.
Cheers,
Magnus
TV
Tom Van Baak
Tue, Dec 4, 2012 1:11 PM
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.
Many Thanks,
Paul
Hi Paul,
-
If you are making frequency measurements, the warm-up of the internal oscillator is the major factor limiting accuracy. This doesn't mean you have wait 12 hours. For example, if all you need is 6 digits of precision, a one-minute warm-up may be sufficient. For each further digit of precision you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It depends on the oscillator. Plotting digits of precision as a function of warm-up time would make a very educational graph you could tape to the top of your SR620.
-
If you are making time interval measurements and using an external standard, the warm-up time will also affect the accuracy of your TI measurements, but to a far lesser degree. Here are informal results for TI (time interval) mode after a 5 minute power-down (see also attached plots):
- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a different counter.
Given that you plan to use the SR620 with high-end GPS gear I would suggest you try this quick experiment for yourself to see what your SR620 does, with your inputs, in your environment. Your numbers will come out different than mine; but the methodology is the same. Your procedures can then be based on measurement and confidence instead of guesswork and folklore.
Note also if your data collection is automated, there's really no reason to wait after power-up at all. Just collect data as soon as you can and skip the predetermined number of samples. I can send you the SR620 GPIB scripts I used for my test. This way you have a record of the warm-up settling time itself, which further gives you confidence in the data that follows.
/tvb
> We are using the SR620 to measure the interval between 1PPS signals from
> two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
> Rubidium clock.
>
> Many Thanks,
> Paul
Hi Paul,
1) If you are making frequency measurements, the warm-up of the internal oscillator is the major factor limiting accuracy. This doesn't mean you have wait 12 hours. For example, if all you need is 6 digits of precision, a one-minute warm-up may be sufficient. For each further digit of precision you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It depends on the oscillator. Plotting digits of precision as a function of warm-up time would make a very educational graph you could tape to the top of your SR620.
2) If you are making time interval measurements and using an external standard, the warm-up time will also affect the accuracy of your TI measurements, but to a far lesser degree. Here are informal results for TI (time interval) mode after a 5 minute power-down (see also attached plots):
- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a different counter.
Given that you plan to use the SR620 with high-end GPS gear I would suggest you try this quick experiment for yourself to see what *your* SR620 does, with *your* inputs, in *your* environment. Your numbers will come out different than mine; but the methodology is the same. Your procedures can then be based on measurement and confidence instead of guesswork and folklore.
Note also if your data collection is automated, there's really no reason to wait after power-up at all. Just collect data as soon as you can and skip the predetermined number of samples. I can send you the SR620 GPIB scripts I used for my test. This way you have a record of the warm-up settling time itself, which further gives you confidence in the data that follows.
/tvb
VE
Volker Esper
Tue, Dec 4, 2012 8:14 PM
Tom,
I agree. Since Paul want's to use an SR620 I presume he needs precision.
Otherwise almost any TIC with a fairly stable osc would do, for example
one with a battery backup. So I further presume that he needs nearly the
full accuracy / stability. But that's just speculation, surely Paul can
answere this question?
Volker
Am 04.12.2012 14:11, schrieb Tom Van Baak:
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.
Many Thanks,
Paul
Hi Paul,
-
If you are making frequency measurements, the warm-up of the internal oscillator is the major factor limiting accuracy. This doesn't mean you have wait 12 hours. For example, if all you need is 6 digits of precision, a one-minute warm-up may be sufficient. For each further digit of precision you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It depends on the oscillator. Plotting digits of precision as a function of warm-up time would make a very educational graph you could tape to the top of your SR620.
-
If you are making time interval measurements and using an external standard, the warm-up time will also affect the accuracy of your TI measurements, but to a far lesser degree. Here are informal results for TI (time interval) mode after a 5 minute power-down (see also attached plots):
- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a different counter.
Given that you plan to use the SR620 with high-end GPS gear I would suggest you try this quick experiment for yourself to see what your SR620 does, with your inputs, in your environment. Your numbers will come out different than mine; but the methodology is the same. Your procedures can then be based on measurement and confidence instead of guesswork and folklore.
Note also if your data collection is automated, there's really no reason to wait after power-up at all. Just collect data as soon as you can and skip the predetermined number of samples. I can send you the SR620 GPIB scripts I used for my test. This way you have a record of the warm-up settling time itself, which further gives you confidence in the data that follows.
/tvb
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Tom,
I agree. Since Paul want's to use an SR620 I presume he needs precision.
Otherwise almost any TIC with a fairly stable osc would do, for example
one with a battery backup. So I further presume that he needs nearly the
full accuracy / stability. But that's just speculation, surely Paul can
answere this question?
Volker
Am 04.12.2012 14:11, schrieb Tom Van Baak:
>> We are using the SR620 to measure the interval between 1PPS signals from
>> two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
>> Rubidium clock.
>>
>> Many Thanks,
>> Paul
>
> Hi Paul,
>
> 1) If you are making frequency measurements, the warm-up of the internal oscillator is the major factor limiting accuracy. This doesn't mean you have wait 12 hours. For example, if all you need is 6 digits of precision, a one-minute warm-up may be sufficient. For each further digit of precision you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It depends on the oscillator. Plotting digits of precision as a function of warm-up time would make a very educational graph you could tape to the top of your SR620.
>
> 2) If you are making time interval measurements and using an external standard, the warm-up time will also affect the accuracy of your TI measurements, but to a far lesser degree. Here are informal results for TI (time interval) mode after a 5 minute power-down (see also attached plots):
>
> - if you need 1 ns accuracy, you can use the SR620 immediately after power-up
> - if you need 100 ps accuracy, wait 2+ minutes
> - if you need 10 ps accuracy, wait 15+ minutes
> - if you need 1 ps accuracy, you need a seriously stable lab environment or a different counter.
>
> Given that you plan to use the SR620 with high-end GPS gear I would suggest you try this quick experiment for yourself to see what *your* SR620 does, with *your* inputs, in *your* environment. Your numbers will come out different than mine; but the methodology is the same. Your procedures can then be based on measurement and confidence instead of guesswork and folklore.
>
> Note also if your data collection is automated, there's really no reason to wait after power-up at all. Just collect data as soon as you can and skip the predetermined number of samples. I can send you the SR620 GPIB scripts I used for my test. This way you have a record of the warm-up settling time itself, which further gives you confidence in the data that follows.
>
> /tvb
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
EG
Eric Garner
Tue, Dec 4, 2012 8:21 PM
I would be interested in those scripts as well. as soon as the replacement
TCXO for my sr620 arrives I can put it back together and this would be fun
to try.
-eric
On Tue, Dec 4, 2012 at 5:11 AM, Tom Van Baak tvb@leapsecond.com wrote:
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is
Rubidium clock.
Many Thanks,
Paul
Hi Paul,
-
If you are making frequency measurements, the warm-up of the internal
oscillator is the major factor limiting accuracy. This doesn't mean you
have wait 12 hours. For example, if all you need is 6 digits of precision,
a one-minute warm-up may be sufficient. For each further digit of precision
you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It
depends on the oscillator. Plotting digits of precision as a function of
warm-up time would make a very educational graph you could tape to the top
of your SR620.
-
If you are making time interval measurements and using an external
standard, the warm-up time will also affect the accuracy of your TI
measurements, but to a far lesser degree. Here are informal results for TI
(time interval) mode after a 5 minute power-down (see also attached plots):
- if you need 1 ns accuracy, you can use the SR620 immediately after
power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment
or a different counter.
Given that you plan to use the SR620 with high-end GPS gear I would
suggest you try this quick experiment for yourself to see what your SR620
does, with your inputs, in your environment. Your numbers will come out
different than mine; but the methodology is the same. Your procedures can
then be based on measurement and confidence instead of guesswork and
folklore.
Note also if your data collection is automated, there's really no reason
to wait after power-up at all. Just collect data as soon as you can and
skip the predetermined number of samples. I can send you the SR620 GPIB
scripts I used for my test. This way you have a record of the warm-up
settling time itself, which further gives you confidence in the data that
follows.
/tvb
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
I would be interested in those scripts as well. as soon as the replacement
TCXO for my sr620 arrives I can put it back together and this would be fun
to try.
-eric
On Tue, Dec 4, 2012 at 5:11 AM, Tom Van Baak <tvb@leapsecond.com> wrote:
> > We are using the SR620 to measure the interval between 1PPS signals from
> > two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is
> a
> > Rubidium clock.
> >
> > Many Thanks,
> > Paul
>
> Hi Paul,
>
> 1) If you are making frequency measurements, the warm-up of the internal
> oscillator is the major factor limiting accuracy. This doesn't mean you
> have wait 12 hours. For example, if all you need is 6 digits of precision,
> a one-minute warm-up may be sufficient. For each further digit of precision
> you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It
> depends on the oscillator. Plotting digits of precision as a function of
> warm-up time would make a very educational graph you could tape to the top
> of your SR620.
>
> 2) If you are making time interval measurements and using an external
> standard, the warm-up time will also affect the accuracy of your TI
> measurements, but to a far lesser degree. Here are informal results for TI
> (time interval) mode after a 5 minute power-down (see also attached plots):
>
> - if you need 1 ns accuracy, you can use the SR620 immediately after
> power-up
> - if you need 100 ps accuracy, wait 2+ minutes
> - if you need 10 ps accuracy, wait 15+ minutes
> - if you need 1 ps accuracy, you need a seriously stable lab environment
> or a different counter.
>
> Given that you plan to use the SR620 with high-end GPS gear I would
> suggest you try this quick experiment for yourself to see what *your* SR620
> does, with *your* inputs, in *your* environment. Your numbers will come out
> different than mine; but the methodology is the same. Your procedures can
> then be based on measurement and confidence instead of guesswork and
> folklore.
>
> Note also if your data collection is automated, there's really no reason
> to wait after power-up at all. Just collect data as soon as you can and
> skip the predetermined number of samples. I can send you the SR620 GPIB
> scripts I used for my test. This way you have a record of the warm-up
> settling time itself, which further gives you confidence in the data that
> follows.
>
> /tvb
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
--
--Eric
_________________________________________
Eric Garner
PD
Paul DeStefano
Thu, Dec 6, 2012 6:10 PM
On Tuesday, 4 December 2012, Tom Van Baak wrote:
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.
Many Thanks,
Paul
-
If you are making frequency measurements, the warm-up of the internal
oscillator is the major factor limiting accuracy. ... Plotting digits of
precision as a function of warm-up time would make a very educational
graph you could tape to the top of your SR620.
-
If you are making time interval measurements and using an external
standard, the warm-up time will also affect the accuracy of your TI
measurements, but to a far lesser degree. Here are informal results for
TI (time interval) mode after a 5 minute power-down (see also attached
plots):
- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a different counter.
Given that you plan to use the SR620 with high-end GPS gear I would
suggest you try this quick experiment for yourself to see what your
SR620 does, with your inputs, in your environment. Your numbers will
come out different than mine; but the methodology is the same. Your
procedures can then be based on measurement and confidence instead of
guesswork and folklore.
Tom & Co.,
Thank you! These plots are excellent and will be very helpful.
You are quite right; we should do the test ourselves. We will definitely
do that. Obviously, there is not need to worry, as we can characterize
the instrument behavior ourselves, which is probably necessary anyway if
we're going to publish these measurements with error values.
Many Thanks,
Paul
--
Paul DeStefano
On Tuesday, 4 December 2012, Tom Van Baak wrote:
>> We are using the SR620 to measure the interval between 1PPS signals from
>> two clocks. One is the Septentrio PolaRx4 GPS receiver and the other is a
>> Rubidium clock.
>>
>> Many Thanks,
>> Paul
> 1) If you are making frequency measurements, the warm-up of the internal
> oscillator is the major factor limiting accuracy. ... Plotting digits of
> precision as a function of warm-up time would make a very educational
> graph you could tape to the top of your SR620.
>
> 2) If you are making time interval measurements and using an external
> standard, the warm-up time will also affect the accuracy of your TI
> measurements, but to a far lesser degree. Here are informal results for
> TI (time interval) mode after a 5 minute power-down (see also attached
> plots):
>
> - if you need 1 ns accuracy, you can use the SR620 immediately after power-up
> - if you need 100 ps accuracy, wait 2+ minutes
> - if you need 10 ps accuracy, wait 15+ minutes
> - if you need 1 ps accuracy, you need a seriously stable lab environment or a different counter.
>
> Given that you plan to use the SR620 with high-end GPS gear I would
> suggest you try this quick experiment for yourself to see what *your*
> SR620 does, with *your* inputs, in *your* environment. Your numbers will
> come out different than mine; but the methodology is the same. Your
> procedures can then be based on measurement and confidence instead of
> guesswork and folklore.
Tom & Co.,
Thank you! These plots are excellent and will be very helpful.
You are quite right; we should do the test ourselves. We will definitely
do that. Obviously, there is not need to worry, as we can characterize
the instrument behavior ourselves, which is probably necessary anyway if
we're going to publish these measurements with error values.
Many Thanks,
Paul
--
Paul DeStefano
PD
Paul DeStefano
Thu, Dec 6, 2012 6:24 PM
Volker,
That's a great question and I'm afraid I don't have a good answer
for you. If pressed, I would estimate less than 100ps.
The error of this measurement contributes to the error in our
final measurement which has many components. I haven't worked out an
error budget for each contributor. Our goal is a final error of less than
10ns. But a few contributors are already expected to be around 1ns, so
this error really needs to be less than 1ns, preferably less than 100ps.
Cheers,
Paul
On Tuesday, 4 December 2012, Volker Esper wrote:
I agree. Since Paul want's to use an SR620 I presume he needs precision.
Otherwise almost any TIC with a fairly stable osc would do, for example
one with a battery backup. So I further presume that he needs nearly the
full accuracy / stability. But that's just speculation, surely Paul can
answere this question?
Volker,
That's a great question and I'm afraid I don't have a good answer
for you. If pressed, I would estimate less than 100ps.
The error of this measurement contributes to the error in our
final measurement which has many components. I haven't worked out an
error budget for each contributor. Our goal is a final error of less than
10ns. But a few contributors are already expected to be around 1ns, so
this error really needs to be less than 1ns, preferably less than 100ps.
Cheers,
Paul
On Tuesday, 4 December 2012, Volker Esper wrote:
> I agree. Since Paul want's to use an SR620 I presume he needs precision.
> Otherwise almost any TIC with a fairly stable osc would do, for example
> one with a battery backup. So I further presume that he needs nearly the
> full accuracy / stability. But that's just speculation, surely Paul can
> answere this question?
--
Paul DeStefano
VE
Volker Esper
Mon, Dec 17, 2012 4:07 PM
I'm curious, which way you went and which accuracy you achieved... :-)
Can you tell us?
Volker
Am 06.12.2012 19:10, schrieb Paul DeStefano:
On Tuesday, 4 December 2012, Tom Van Baak wrote:
We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other
is a
Rubidium clock.
Many Thanks,
Paul
-
If you are making frequency measurements, the warm-up of the
internal oscillator is the major factor limiting accuracy. ...
Plotting digits of precision as a function of warm-up time would make
a very educational graph you could tape to the top of your SR620.
-
If you are making time interval measurements and using an external
standard, the warm-up time will also affect the accuracy of your TI
measurements, but to a far lesser degree. Here are informal results
for TI (time interval) mode after a 5 minute power-down (see also
attached plots):
- if you need 1 ns accuracy, you can use the SR620 immediately after
power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab
environment or a different counter.
Given that you plan to use the SR620 with high-end GPS gear I would
suggest you try this quick experiment for yourself to see what your
SR620 does, with your inputs, in your environment. Your numbers
will come out different than mine; but the methodology is the same.
Your procedures can then be based on measurement and confidence
instead of guesswork and folklore.
Tom & Co.,
Thank you! These plots are excellent and will be very helpful. You are
quite right; we should do the test ourselves. We will definitely do
that. Obviously, there is not need to worry, as we can characterize the
instrument behavior ourselves, which is probably necessary anyway if
we're going to publish these measurements with error values.
Many Thanks,
Paul
I'm curious, which way you went and which accuracy you achieved... :-)
Can you tell us?
Volker
Am 06.12.2012 19:10, schrieb Paul DeStefano:
> On Tuesday, 4 December 2012, Tom Van Baak wrote:
>>> We are using the SR620 to measure the interval between 1PPS signals from
>>> two clocks. One is the Septentrio PolaRx4 GPS receiver and the other
>>> is a
>>> Rubidium clock.
>>>
>>> Many Thanks,
>>> Paul
>> 1) If you are making frequency measurements, the warm-up of the
>> internal oscillator is the major factor limiting accuracy. ...
>> Plotting digits of precision as a function of warm-up time would make
>> a very educational graph you could tape to the top of your SR620.
>>
>> 2) If you are making time interval measurements and using an external
>> standard, the warm-up time will also affect the accuracy of your TI
>> measurements, but to a far lesser degree. Here are informal results
>> for TI (time interval) mode after a 5 minute power-down (see also
>> attached plots):
>>
>> - if you need 1 ns accuracy, you can use the SR620 immediately after
>> power-up
>> - if you need 100 ps accuracy, wait 2+ minutes
>> - if you need 10 ps accuracy, wait 15+ minutes
>> - if you need 1 ps accuracy, you need a seriously stable lab
>> environment or a different counter.
>>
>> Given that you plan to use the SR620 with high-end GPS gear I would
>> suggest you try this quick experiment for yourself to see what *your*
>> SR620 does, with *your* inputs, in *your* environment. Your numbers
>> will come out different than mine; but the methodology is the same.
>> Your procedures can then be based on measurement and confidence
>> instead of guesswork and folklore.
>
> Tom & Co.,
>
> Thank you! These plots are excellent and will be very helpful. You are
> quite right; we should do the test ourselves. We will definitely do
> that. Obviously, there is not need to worry, as we can characterize the
> instrument behavior ourselves, which is probably necessary anyway if
> we're going to publish these measurements with error values.
>
> Many Thanks,
> Paul
>