SK
Stan Katz
Wed, Mar 11, 2015 5:42 AM
Now that I've restored my HP419A to running order for the meantime, I have
nagging questions regarding the continuous use of the attenuator switch.
The main board of the instrument, A4, was badly polluted with an invisible
layer of potassium hydroxide from the 30 years of being in the same box
with dead/mummified nicad batteries.
In accordance with
http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf I
prepared a 3% boric acid solution and flushed the affected circuit board
several times, while scrubbing with a soft nylon toothbrush. I then flushed
the board with hot water. After leaving the board on my furnace plenum for
a few days to thoroughly dry out, the board problem was sorted out. If the
KOH deposited an invisible layer on A4, then certainly, there must be a
layer of KOH on the attenuator switch's gold contacts, as well. I posted my
concern about the KOH on a premier chemistry forum. The administrator
himself answered my query. He indicated that the only way to remove the KOH
from the switch is to thoroughly flush with di water. There's no place for
a flushing stream of water to exit the instrument safely. I'm not about
to remove the switch! Now, I don't know the hardness of KOH, but Potassium
has a very small .5 mohs of hardness. I presume, that KOH is similar. In
any event, I would tend to believe there will be some small abrasive wear
from the KOH. The attenuator on a null meter is heavily utilized, and even
a light abrasive can cause trouble eventually. Proper gold plating, that's
kept clean, never suffers from abrasive wear. Improper plating, can suffer
from adhesive wear, but I have confidence that the HP of the 1960s-80s was
using the finest gold plating available. I thought about applying one the
the Deoxit products. A careful reading of the Deoxit literature claims that
the D-series treatment is for "non-critical metal surfaces". I think that
the gold plating in the HP419A qualifies as "quite critical". The Deoxit G
series literature leads me to believe it may not have enough cleaning
action to flush out the KOH. Also, please note, only the removal of oxides
is mentioned as a feature of the Deoxit line. What other corrosion products
that may be removed are not mentioned.
A proper "science project" would be to obtain Bromothymol blue indicator
solution, and prepare it with Phosphoric acid in accordance with the above
citation. I could then apply it to the switch contacts ( in itself a PITA
job ), and then try the Deoxit G series, and perhaps other brands of
contact cleaner to see if the KOH is truly removed without resorting to
water. I'm actually willing to take on such a project ( it's a hobby, after
all ), but I don't even know how to procure a proper concentration of
Bromothymol blue from over the internet. Without access to purchase orders
through a university, or corporation, even innocuous chemicals are
unobtainable from US chemical supply houses directly by individuals. There
are lots of offers on Ebay, but I have no way of knowing if what's sold is
a the proper concentration. I don't even know how to specify what that
concentration should be.
Yes, I should get on with my life, but I've invested quite a few hours
restoring this instrument, and I just can't let this situation fester
without a resolution. Can anyone on the list help put me out of my misery?
Stan
Now that I've restored my HP419A to running order for the meantime, I have
nagging questions regarding the continuous use of the attenuator switch.
The main board of the instrument, A4, was badly polluted with an invisible
layer of potassium hydroxide from the 30 years of being in the same box
with dead/mummified nicad batteries.
In accordance with
http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf I
prepared a 3% boric acid solution and flushed the affected circuit board
several times, while scrubbing with a soft nylon toothbrush. I then flushed
the board with hot water. After leaving the board on my furnace plenum for
a few days to thoroughly dry out, the board problem was sorted out. If the
KOH deposited an invisible layer on A4, then certainly, there must be a
layer of KOH on the attenuator switch's gold contacts, as well. I posted my
concern about the KOH on a premier chemistry forum. The administrator
himself answered my query. He indicated that the only way to remove the KOH
from the switch is to thoroughly flush with di water. There's no place for
a flushing stream of water to exit the instrument safely. I'm not about
to remove the switch! Now, I don't know the hardness of KOH, but Potassium
has a very small .5 mohs of hardness. I presume, that KOH is similar. In
any event, I would tend to believe there will be some small abrasive wear
from the KOH. The attenuator on a null meter is heavily utilized, and even
a light abrasive can cause trouble eventually. Proper gold plating, that's
kept clean, never suffers from abrasive wear. Improper plating, can suffer
from adhesive wear, but I have confidence that the HP of the 1960s-80s was
using the finest gold plating available. I thought about applying one the
the Deoxit products. A careful reading of the Deoxit literature claims that
the D-series treatment is for "non-critical metal surfaces". I think that
the gold plating in the HP419A qualifies as "quite critical". The Deoxit G
series literature leads me to believe it may not have enough cleaning
action to flush out the KOH. Also, please note, only the removal of oxides
is mentioned as a feature of the Deoxit line. What other corrosion products
that may be removed are not mentioned.
A proper "science project" would be to obtain Bromothymol blue indicator
solution, and prepare it with Phosphoric acid in accordance with the above
citation. I could then apply it to the switch contacts ( in itself a PITA
job ), and then try the Deoxit G series, and perhaps other brands of
contact cleaner to see if the KOH is truly removed without resorting to
water. I'm actually willing to take on such a project ( it's a hobby, after
all ), but I don't even know how to procure a proper concentration of
Bromothymol blue from over the internet. Without access to purchase orders
through a university, or corporation, even innocuous chemicals are
unobtainable from US chemical supply houses directly by individuals. There
are lots of offers on Ebay, but I have no way of knowing if what's sold is
a the proper concentration. I don't even know how to specify what that
concentration should be.
Yes, I should get on with my life, but I've invested quite a few hours
restoring this instrument, and I just can't let this situation fester
without a resolution. Can anyone on the list help put me out of my misery?
Stan
CH
Chuck Harris
Wed, Mar 11, 2015 6:26 AM
Does KOH really evaporate, and spread in the wind?
I don't think it does.
In my experience, the damage done by nicad's is limited to
where the electrolyte dribbles, spatters, or wicks.
Powdered electrolyte can get all over things, but usually
it can be just brushed away, as it is not too corrosive
when dry.
If the switches are not covered with KOH dust and crust, I
truly doubt that they have any measurable amount of KOH
inside of them.
However,
If you want to be really sure, go to your grocery store and
buy yourself a gallon of steam distilled water for $0.50, or
so, and an empty spray bottle, like is used for window cleaner
and liberally hose out the switches and any other area you
are concerned about.
When you are done, shake, blow, or simply let drip as much
water as you can off of the unit. Set the unit up next to
a fan, and let the fan blow air on it for a couple of days.
It will dry out.
If you want to be doubly sure, a convection oven is the
answer. There are several ways you can get one on the cheap.
If your house has an electric convection oven in the kitchen,
you can use that set to 150F (50-60C). Or, you can put a
muffin fan on a simple stand, and put it in an electric oven,
with the door closed, and the oven light turned on. (Gas
ovens are fine too, if they do not have a pilot light.)
And, if you want to stay away from the kitchen, you can fashion
a simple convection oven using a cardboard box, a 40W lamp,
and a small muffin fan. Make sure the box is big enough so
that the lamp is no closer than 12 inches from any side of the
box, and make sure that the fan and lamp cannot fall over, or
the box move and cause issues... don't want to make a fire
hazard.
-Chuck Harris
Stan Katz wrote:
Now that I've restored my HP419A to running order for the meantime, I have
nagging questions regarding the continuous use of the attenuator switch.
The main board of the instrument, A4, was badly polluted with an invisible
layer of potassium hydroxide from the 30 years of being in the same box
with dead/mummified nicad batteries.
In accordance with
http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf I
prepared a 3% boric acid solution and flushed the affected circuit board
several times, while scrubbing with a soft nylon toothbrush. I then flushed
the board with hot water. After leaving the board on my furnace plenum for
a few days to thoroughly dry out, the board problem was sorted out. If the
KOH deposited an invisible layer on A4, then certainly, there must be a
layer of KOH on the attenuator switch's gold contacts, as well. I posted my
concern about the KOH on a premier chemistry forum. The administrator
himself answered my query. He indicated that the only way to remove the KOH
from the switch is to thoroughly flush with di water. There's no place for
a flushing stream of water to exit the instrument safely. I'm not about
to remove the switch! Now, I don't know the hardness of KOH, but Potassium
has a very small .5 mohs of hardness. I presume, that KOH is similar. In
any event, I would tend to believe there will be some small abrasive wear
from the KOH. The attenuator on a null meter is heavily utilized, and even
a light abrasive can cause trouble eventually. Proper gold plating, that's
kept clean, never suffers from abrasive wear. Improper plating, can suffer
from adhesive wear, but I have confidence that the HP of the 1960s-80s was
using the finest gold plating available. I thought about applying one the
the Deoxit products. A careful reading of the Deoxit literature claims that
the D-series treatment is for "non-critical metal surfaces". I think that
the gold plating in the HP419A qualifies as "quite critical". The Deoxit G
series literature leads me to believe it may not have enough cleaning
action to flush out the KOH. Also, please note, only the removal of oxides
is mentioned as a feature of the Deoxit line. What other corrosion products
that may be removed are not mentioned.
A proper "science project" would be to obtain Bromothymol blue indicator
solution, and prepare it with Phosphoric acid in accordance with the above
citation. I could then apply it to the switch contacts ( in itself a PITA
job ), and then try the Deoxit G series, and perhaps other brands of
contact cleaner to see if the KOH is truly removed without resorting to
water. I'm actually willing to take on such a project ( it's a hobby, after
all ), but I don't even know how to procure a proper concentration of
Bromothymol blue from over the internet. Without access to purchase orders
through a university, or corporation, even innocuous chemicals are
unobtainable from US chemical supply houses directly by individuals. There
are lots of offers on Ebay, but I have no way of knowing if what's sold is
a the proper concentration. I don't even know how to specify what that
concentration should be.
Yes, I should get on with my life, but I've invested quite a few hours
restoring this instrument, and I just can't let this situation fester
without a resolution. Can anyone on the list help put me out of my misery?
Stan
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Does KOH really evaporate, and spread in the wind?
I don't think it does.
In my experience, the damage done by nicad's is limited to
where the electrolyte dribbles, spatters, or wicks.
Powdered electrolyte can get all over things, but usually
it can be just brushed away, as it is not too corrosive
when dry.
If the switches are not covered with KOH dust and crust, I
truly doubt that they have any measurable amount of KOH
inside of them.
However,
If you want to be really sure, go to your grocery store and
buy yourself a gallon of steam distilled water for $0.50, or
so, and an empty spray bottle, like is used for window cleaner
and liberally hose out the switches and any other area you
are concerned about.
When you are done, shake, blow, or simply let drip as much
water as you can off of the unit. Set the unit up next to
a fan, and let the fan blow air on it for a couple of days.
It will dry out.
If you want to be doubly sure, a convection oven is the
answer. There are several ways you can get one on the cheap.
If your house has an electric convection oven in the kitchen,
you can use that set to 150F (50-60C). Or, you can put a
muffin fan on a simple stand, and put it in an electric oven,
with the door closed, and the oven light turned on. (Gas
ovens are fine too, if they do not have a pilot light.)
And, if you want to stay away from the kitchen, you can fashion
a simple convection oven using a cardboard box, a 40W lamp,
and a small muffin fan. Make sure the box is big enough so
that the lamp is no closer than 12 inches from any side of the
box, and make sure that the fan and lamp cannot fall over, or
the box move and cause issues... don't want to make a fire
hazard.
-Chuck Harris
Stan Katz wrote:
> Now that I've restored my HP419A to running order for the meantime, I have
> nagging questions regarding the continuous use of the attenuator switch.
> The main board of the instrument, A4, was badly polluted with an invisible
> layer of potassium hydroxide from the 30 years of being in the same box
> with dead/mummified nicad batteries.
>
> In accordance with
> http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf I
> prepared a 3% boric acid solution and flushed the affected circuit board
> several times, while scrubbing with a soft nylon toothbrush. I then flushed
> the board with hot water. After leaving the board on my furnace plenum for
> a few days to thoroughly dry out, the board problem was sorted out. If the
> KOH deposited an invisible layer on A4, then certainly, there must be a
> layer of KOH on the attenuator switch's gold contacts, as well. I posted my
> concern about the KOH on a premier chemistry forum. The administrator
> himself answered my query. He indicated that the only way to remove the KOH
> from the switch is to thoroughly flush with di water. There's no place for
> a flushing stream of water to exit the instrument safely. I'm not about
> to remove the switch! Now, I don't know the hardness of KOH, but Potassium
> has a very small .5 mohs of hardness. I presume, that KOH is similar. In
> any event, I would tend to believe there will be some small abrasive wear
> from the KOH. The attenuator on a null meter is heavily utilized, and even
> a light abrasive can cause trouble eventually. Proper gold plating, that's
> kept clean, never suffers from abrasive wear. Improper plating, can suffer
> from adhesive wear, but I have confidence that the HP of the 1960s-80s was
> using the finest gold plating available. I thought about applying one the
> the Deoxit products. A careful reading of the Deoxit literature claims that
> the D-series treatment is for "non-critical metal surfaces". I think that
> the gold plating in the HP419A qualifies as "quite critical". The Deoxit G
> series literature leads me to believe it may not have enough cleaning
> action to flush out the KOH. Also, please note, only the removal of oxides
> is mentioned as a feature of the Deoxit line. What other corrosion products
> that may be removed are not mentioned.
>
> A proper "science project" would be to obtain Bromothymol blue indicator
> solution, and prepare it with Phosphoric acid in accordance with the above
> citation. I could then apply it to the switch contacts ( in itself a PITA
> job ), and then try the Deoxit G series, and perhaps other brands of
> contact cleaner to see if the KOH is truly removed without resorting to
> water. I'm actually willing to take on such a project ( it's a hobby, after
> all ), but I don't even know how to procure a proper concentration of
> Bromothymol blue from over the internet. Without access to purchase orders
> through a university, or corporation, even innocuous chemicals are
> unobtainable from US chemical supply houses directly by individuals. There
> are lots of offers on Ebay, but I have no way of knowing if what's sold is
> a the proper concentration. I don't even know how to specify what that
> concentration should be.
>
> Yes, I should get on with my life, but I've invested quite a few hours
> restoring this instrument, and I just can't let this situation fester
> without a resolution. Can anyone on the list help put me out of my misery?
>
> Stan
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
MK
M K
Wed, Mar 11, 2015 7:03 AM
On 11/03/2015 06:26, Chuck Harris wrote:
Does KOH really evaporate, and spread in the wind?
I don't think it does.
In my experience, the damage done by nicad's is limited to
where the electrolyte dribbles, spatters, or wicks.
Powdered electrolyte can get all over things, but usually
it can be just brushed away, as it is not too corrosive
when dry.
If the switches are not covered with KOH dust and crust, I
truly doubt that they have any measurable amount of KOH
inside of them.
However,
If you want to be really sure, go to your grocery store and
buy yourself a gallon of steam distilled water for $0.50, or
so, and an empty spray bottle, like is used for window cleaner
and liberally hose out the switches and any other area you
are concerned about.
When you are done, shake, blow, or simply let drip as much
water as you can off of the unit. Set the unit up next to
a fan, and let the fan blow air on it for a couple of days.
It will dry out.
If you want to be doubly sure, a convection oven is the
answer. There are several ways you can get one on the cheap.
If your house has an electric convection oven in the kitchen,
you can use that set to 150F (50-60C). Or, you can put a
muffin fan on a simple stand, and put it in an electric oven,
with the door closed, and the oven light turned on. (Gas
ovens are fine too, if they do not have a pilot light.)
And, if you want to stay away from the kitchen, you can fashion
a simple convection oven using a cardboard box, a 40W lamp,
and a small muffin fan. Make sure the box is big enough so
that the lamp is no closer than 12 inches from any side of the
box, and make sure that the fan and lamp cannot fall over, or
the box move and cause issues... don't want to make a fire
hazard.
-Chuck Harris
Stan Katz wrote:
Now that I've restored my HP419A to running order for the meantime, I
have
nagging questions regarding the continuous use of the attenuator switch.
The main board of the instrument, A4, was badly polluted with an
invisible
layer of potassium hydroxide from the 30 years of being in the same box
with dead/mummified nicad batteries.
In accordance with
http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf
I
prepared a 3% boric acid solution and flushed the affected circuit board
several times, while scrubbing with a soft nylon toothbrush. I then
flushed
the board with hot water. After leaving the board on my furnace
plenum for
a few days to thoroughly dry out, the board problem was sorted out.
If the
KOH deposited an invisible layer on A4, then certainly, there must be a
layer of KOH on the attenuator switch's gold contacts, as well. I
posted my
concern about the KOH on a premier chemistry forum. The administrator
himself answered my query. He indicated that the only way to remove
the KOH
from the switch is to thoroughly flush with di water. There's no
place for
a flushing stream of water to exit the instrument safely. I'm not
about
to remove the switch! Now, I don't know the hardness of KOH, but
Potassium
has a very small .5 mohs of hardness. I presume, that KOH is similar. In
any event, I would tend to believe there will be some small abrasive
wear
from the KOH. The attenuator on a null meter is heavily utilized, and
even
a light abrasive can cause trouble eventually. Proper gold plating,
that's
kept clean, never suffers from abrasive wear. Improper plating, can
suffer
from adhesive wear, but I have confidence that the HP of the
1960s-80s was
using the finest gold plating available. I thought about applying
one the
the Deoxit products. A careful reading of the Deoxit literature
claims that
the D-series treatment is for "non-critical metal surfaces". I think
that
the gold plating in the HP419A qualifies as "quite critical". The
Deoxit G
series literature leads me to believe it may not have enough cleaning
action to flush out the KOH. Also, please note, only the removal of
oxides
is mentioned as a feature of the Deoxit line. What other corrosion
products
that may be removed are not mentioned.
A proper "science project" would be to obtain Bromothymol blue indicator
solution, and prepare it with Phosphoric acid in accordance with the
above
citation. I could then apply it to the switch contacts ( in itself a
PITA
job ), and then try the Deoxit G series, and perhaps other brands of
contact cleaner to see if the KOH is truly removed without resorting to
water. I'm actually willing to take on such a project ( it's a hobby,
after
all ), but I don't even know how to procure a proper concentration of
Bromothymol blue from over the internet. Without access to purchase
orders
through a university, or corporation, even innocuous chemicals are
unobtainable from US chemical supply houses directly by individuals.
There
are lots of offers on Ebay, but I have no way of knowing if what's
sold is
a the proper concentration. I don't even know how to specify what that
concentration should be.
Yes, I should get on with my life, but I've invested quite a few hours
restoring this instrument, and I just can't let this situation fester
without a resolution. Can anyone on the list help put me out of my
misery?
Stan
I would have thought that KOH would turn into K2CO3 over time, and that
would not be quite so agressive to deal with
On 11/03/2015 06:26, Chuck Harris wrote:
> Does KOH really evaporate, and spread in the wind?
>
> I don't think it does.
>
> In my experience, the damage done by nicad's is limited to
> where the electrolyte dribbles, spatters, or wicks.
> Powdered electrolyte can get all over things, but usually
> it can be just brushed away, as it is not too corrosive
> when dry.
>
> If the switches are not covered with KOH dust and crust, I
> truly doubt that they have any measurable amount of KOH
> inside of them.
>
> However,
>
> If you want to be really sure, go to your grocery store and
> buy yourself a gallon of steam distilled water for $0.50, or
> so, and an empty spray bottle, like is used for window cleaner
> and liberally hose out the switches and any other area you
> are concerned about.
>
> When you are done, shake, blow, or simply let drip as much
> water as you can off of the unit. Set the unit up next to
> a fan, and let the fan blow air on it for a couple of days.
>
> It will dry out.
>
> If you want to be doubly sure, a convection oven is the
> answer. There are several ways you can get one on the cheap.
> If your house has an electric convection oven in the kitchen,
> you can use that set to 150F (50-60C). Or, you can put a
> muffin fan on a simple stand, and put it in an electric oven,
> with the door closed, and the oven light turned on. (Gas
> ovens are fine too, if they do not have a pilot light.)
>
> And, if you want to stay away from the kitchen, you can fashion
> a simple convection oven using a cardboard box, a 40W lamp,
> and a small muffin fan. Make sure the box is big enough so
> that the lamp is no closer than 12 inches from any side of the
> box, and make sure that the fan and lamp cannot fall over, or
> the box move and cause issues... don't want to make a fire
> hazard.
>
> -Chuck Harris
>
> Stan Katz wrote:
>> Now that I've restored my HP419A to running order for the meantime, I
>> have
>> nagging questions regarding the continuous use of the attenuator switch.
>> The main board of the instrument, A4, was badly polluted with an
>> invisible
>> layer of potassium hydroxide from the 30 years of being in the same box
>> with dead/mummified nicad batteries.
>>
>> In accordance with
>> http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf
>> I
>> prepared a 3% boric acid solution and flushed the affected circuit board
>> several times, while scrubbing with a soft nylon toothbrush. I then
>> flushed
>> the board with hot water. After leaving the board on my furnace
>> plenum for
>> a few days to thoroughly dry out, the board problem was sorted out.
>> If the
>> KOH deposited an invisible layer on A4, then certainly, there must be a
>> layer of KOH on the attenuator switch's gold contacts, as well. I
>> posted my
>> concern about the KOH on a premier chemistry forum. The administrator
>> himself answered my query. He indicated that the only way to remove
>> the KOH
>> from the switch is to thoroughly flush with di water. There's no
>> place for
>> a flushing stream of water to exit the instrument safely. I'm not
>> about
>> to remove the switch! Now, I don't know the hardness of KOH, but
>> Potassium
>> has a very small .5 mohs of hardness. I presume, that KOH is similar. In
>> any event, I would tend to believe there will be some small abrasive
>> wear
>> from the KOH. The attenuator on a null meter is heavily utilized, and
>> even
>> a light abrasive can cause trouble eventually. Proper gold plating,
>> that's
>> kept clean, never suffers from abrasive wear. Improper plating, can
>> suffer
>> from adhesive wear, but I have confidence that the HP of the
>> 1960s-80s was
>> using the finest gold plating available. I thought about applying
>> one the
>> the Deoxit products. A careful reading of the Deoxit literature
>> claims that
>> the D-series treatment is for "non-critical metal surfaces". I think
>> that
>> the gold plating in the HP419A qualifies as "quite critical". The
>> Deoxit G
>> series literature leads me to believe it may not have enough cleaning
>> action to flush out the KOH. Also, please note, only the removal of
>> oxides
>> is mentioned as a feature of the Deoxit line. What other corrosion
>> products
>> that may be removed are not mentioned.
>>
>> A proper "science project" would be to obtain Bromothymol blue indicator
>> solution, and prepare it with Phosphoric acid in accordance with the
>> above
>> citation. I could then apply it to the switch contacts ( in itself a
>> PITA
>> job ), and then try the Deoxit G series, and perhaps other brands of
>> contact cleaner to see if the KOH is truly removed without resorting to
>> water. I'm actually willing to take on such a project ( it's a hobby,
>> after
>> all ), but I don't even know how to procure a proper concentration of
>> Bromothymol blue from over the internet. Without access to purchase
>> orders
>> through a university, or corporation, even innocuous chemicals are
>> unobtainable from US chemical supply houses directly by individuals.
>> There
>> are lots of offers on Ebay, but I have no way of knowing if what's
>> sold is
>> a the proper concentration. I don't even know how to specify what that
>> concentration should be.
>>
>> Yes, I should get on with my life, but I've invested quite a few hours
>> restoring this instrument, and I just can't let this situation fester
>> without a resolution. Can anyone on the list help put me out of my
>> misery?
>>
>> Stan
>> _______________________________________________
>
I would have thought that KOH would turn into K2CO3 over time, and that
would not be quite so agressive to deal with
F
fala@gmx.net
Wed, Mar 11, 2015 7:04 AM
Hello Stan,
since KOH is water soluble, thorough cleaning with a moist cloth (di
water) should do the job as well. Just wear proper protection, i.e.
safety goggles and gloves. I myself would use disposable latex gloves
from the grocery store in this instance, since you are dealing with very
dilute chemicals (beware: latex gloves do not protect against chemicals!
If in doubt, use (disposable) nitrile gloves).
Furthermore, instead of bromthymol blue, you could check your progress
with a "universal indicator" or "pH indicator" paper as well (ref.
Wikipedia "Universal indicator"). Should be easier to obtain.Indicator
paper is a disposable roll of thin paper, treated with several indicator
chemicals (e.g. bromthymol blue and others) which change color based on
pH. After wetting the paper with di water, observe if there is a color
change after contacting the polluted parts (KOH is basic).
Am 11.03.2015 um 06:42 schrieb Stan Katz:
Now that I've restored my HP419A to running order for the meantime, I have
nagging questions regarding the continuous use of the attenuator switch.
The main board of the instrument, A4, was badly polluted with an invisible
layer of potassium hydroxide from the 30 years of being in the same box
with dead/mummified nicad batteries.
In accordance with
http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf I
prepared a 3% boric acid solution and flushed the affected circuit board
several times, while scrubbing with a soft nylon toothbrush. I then flushed
the board with hot water. After leaving the board on my furnace plenum for
a few days to thoroughly dry out, the board problem was sorted out. If the
KOH deposited an invisible layer on A4, then certainly, there must be a
layer of KOH on the attenuator switch's gold contacts, as well. I posted my
concern about the KOH on a premier chemistry forum. The administrator
himself answered my query. He indicated that the only way to remove the KOH
from the switch is to thoroughly flush with di water. There's no place for
a flushing stream of water to exit the instrument safely. I'm not about
to remove the switch! Now, I don't know the hardness of KOH, but Potassium
has a very small .5 mohs of hardness. I presume, that KOH is similar. In
any event, I would tend to believe there will be some small abrasive wear
from the KOH. The attenuator on a null meter is heavily utilized, and even
a light abrasive can cause trouble eventually. Proper gold plating, that's
kept clean, never suffers from abrasive wear. Improper plating, can suffer
from adhesive wear, but I have confidence that the HP of the 1960s-80s was
using the finest gold plating available. I thought about applying one the
the Deoxit products. A careful reading of the Deoxit literature claims that
the D-series treatment is for "non-critical metal surfaces". I think that
the gold plating in the HP419A qualifies as "quite critical". The Deoxit G
series literature leads me to believe it may not have enough cleaning
action to flush out the KOH. Also, please note, only the removal of oxides
is mentioned as a feature of the Deoxit line. What other corrosion products
that may be removed are not mentioned.
A proper "science project" would be to obtain Bromothymol blue indicator
solution, and prepare it with Phosphoric acid in accordance with the above
citation. I could then apply it to the switch contacts ( in itself a PITA
job ), and then try the Deoxit G series, and perhaps other brands of
contact cleaner to see if the KOH is truly removed without resorting to
water. I'm actually willing to take on such a project ( it's a hobby, after
all ), but I don't even know how to procure a proper concentration of
Bromothymol blue from over the internet. Without access to purchase orders
through a university, or corporation, even innocuous chemicals are
unobtainable from US chemical supply houses directly by individuals. There
are lots of offers on Ebay, but I have no way of knowing if what's sold is
a the proper concentration. I don't even know how to specify what that
concentration should be.
Yes, I should get on with my life, but I've invested quite a few hours
restoring this instrument, and I just can't let this situation fester
without a resolution. Can anyone on the list help put me out of my misery?
Stan
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Stan,
since KOH is water soluble, thorough cleaning with a moist cloth (di
water) should do the job as well. Just wear proper protection, i.e.
safety goggles and gloves. I myself would use disposable latex gloves
from the grocery store in this instance, since you are dealing with very
dilute chemicals (beware: latex gloves do not protect against chemicals!
If in doubt, use (disposable) nitrile gloves).
Furthermore, instead of bromthymol blue, you could check your progress
with a "universal indicator" or "pH indicator" paper as well (ref.
Wikipedia "Universal indicator"). Should be easier to obtain.Indicator
paper is a disposable roll of thin paper, treated with several indicator
chemicals (e.g. bromthymol blue and others) which change color based on
pH. After wetting the paper with di water, observe if there is a color
change after contacting the polluted parts (KOH is basic).
- Firat
Am 11.03.2015 um 06:42 schrieb Stan Katz:
> Now that I've restored my HP419A to running order for the meantime, I have
> nagging questions regarding the continuous use of the attenuator switch.
> The main board of the instrument, A4, was badly polluted with an invisible
> layer of potassium hydroxide from the 30 years of being in the same box
> with dead/mummified nicad batteries.
>
> In accordance with
> http://www.faa.gov/documentLibrary/media/Advisory_Circular/AC43-206part2.pdf I
> prepared a 3% boric acid solution and flushed the affected circuit board
> several times, while scrubbing with a soft nylon toothbrush. I then flushed
> the board with hot water. After leaving the board on my furnace plenum for
> a few days to thoroughly dry out, the board problem was sorted out. If the
> KOH deposited an invisible layer on A4, then certainly, there must be a
> layer of KOH on the attenuator switch's gold contacts, as well. I posted my
> concern about the KOH on a premier chemistry forum. The administrator
> himself answered my query. He indicated that the only way to remove the KOH
> from the switch is to thoroughly flush with di water. There's no place for
> a flushing stream of water to exit the instrument safely. I'm not about
> to remove the switch! Now, I don't know the hardness of KOH, but Potassium
> has a very small .5 mohs of hardness. I presume, that KOH is similar. In
> any event, I would tend to believe there will be some small abrasive wear
> from the KOH. The attenuator on a null meter is heavily utilized, and even
> a light abrasive can cause trouble eventually. Proper gold plating, that's
> kept clean, never suffers from abrasive wear. Improper plating, can suffer
> from adhesive wear, but I have confidence that the HP of the 1960s-80s was
> using the finest gold plating available. I thought about applying one the
> the Deoxit products. A careful reading of the Deoxit literature claims that
> the D-series treatment is for "non-critical metal surfaces". I think that
> the gold plating in the HP419A qualifies as "quite critical". The Deoxit G
> series literature leads me to believe it may not have enough cleaning
> action to flush out the KOH. Also, please note, only the removal of oxides
> is mentioned as a feature of the Deoxit line. What other corrosion products
> that may be removed are not mentioned.
>
> A proper "science project" would be to obtain Bromothymol blue indicator
> solution, and prepare it with Phosphoric acid in accordance with the above
> citation. I could then apply it to the switch contacts ( in itself a PITA
> job ), and then try the Deoxit G series, and perhaps other brands of
> contact cleaner to see if the KOH is truly removed without resorting to
> water. I'm actually willing to take on such a project ( it's a hobby, after
> all ), but I don't even know how to procure a proper concentration of
> Bromothymol blue from over the internet. Without access to purchase orders
> through a university, or corporation, even innocuous chemicals are
> unobtainable from US chemical supply houses directly by individuals. There
> are lots of offers on Ebay, but I have no way of knowing if what's sold is
> a the proper concentration. I don't even know how to specify what that
> concentration should be.
>
> Yes, I should get on with my life, but I've invested quite a few hours
> restoring this instrument, and I just can't let this situation fester
> without a resolution. Can anyone on the list help put me out of my misery?
>
> Stan
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
CH
Chuck Harris
Wed, Mar 11, 2015 1:58 PM
That is a predominate reaction, but only on the surface.
Under the carbonate layer, you will find lots of nice
caustic KOH.
When I clean up NiCad vomit, I put the crusted parts into
a cup of white vinegar, where they fizz for hours.
I use white vinegar because it is distilled, and as such,
everything in it evaporates freely. The salts, etc., that
come from this, or that, reacting with the vinegar will
not, evaporate, and need to be rinsed.
I would not, however, expect to find any KOH out in the
outer extremities, where the OP is worried. I wouldn't
expect to find anything but whatever travelled there as
dust.
That dust, I blow away with compressed air.
-Chuck Harris
M K wrote:
That is a predominate reaction, but only on the surface.
Under the carbonate layer, you will find lots of nice
caustic KOH.
When I clean up NiCad vomit, I put the crusted parts into
a cup of white vinegar, where they fizz for hours.
I use white vinegar because it is distilled, and as such,
everything in it evaporates freely. The salts, etc., that
come from this, or that, reacting with the vinegar will
not, evaporate, and need to be rinsed.
I would not, however, expect to find any KOH out in the
outer extremities, where the OP is worried. I wouldn't
expect to find anything but whatever travelled there as
dust.
That dust, I blow away with compressed air.
-Chuck Harris
M K wrote:
>>> _______________________________________________
>>
> I would have thought that KOH would turn into K2CO3 over time, and that would not be
> quite so agressive to deal with
> _______________________________________________