John wrote:
David mentioned this:
* * *
What do people think of the device?
What are you trying to accomplish? As I posted previously:
Let's assume that it is still working exactly the same as it was
when it was calibrated. The calibration values are recorded to
4-1/2 digits. So the uncertainty is greater than the two LSDs of a
3457A. Of course, it is almost certainly not working exactly the
same as it was when it was calibrated, [so its last digit is going
to be uncertain,] which is going to put at least the third LSD of a
3457A in question. So it is good enough for a 3457A only if one is
content with a 3-1/2 digit verification on a 6-1/2 digit instrument.
You should be able to rely on it to verify instruments to 3-1/2
digits. Beyond that, you would need something with a smaller uncertainty.
Best regards,
Charles
My main interest is as a 'sanity check'. I use DMMs for engineering purposes.
Shields up:
IMO, there are very few applications, other than fundamental physics
research, that really NEED 5,6,7, or 8+ digits.
I would no more use a multi digit DVM to do what can be done with a
differential voltmeter than use a carrier phase tracking GPS to measure
for a living room carpet.
Shields down.
-John
================
John wrote:
David mentioned this:
* * *
What do people think of the device?
What are you trying to accomplish? As I posted previously:
Let's assume that it is still working exactly the same as it was
when it was calibrated. The calibration values are recorded to
4-1/2 digits. So the uncertainty is greater than the two LSDs of a
3457A. Of course, it is almost certainly not working exactly the
same as it was when it was calibrated, [so its last digit is going
to be uncertain,] which is going to put at least the third LSD of a
3457A in question. So it is good enough for a 3457A only if one is
content with a 3-1/2 digit verification on a 6-1/2 digit instrument.
You should be able to rely on it to verify instruments to 3-1/2
digits. Beyond that, you would need something with a smaller uncertainty.
Best regards,
Charles
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
As digits rise, applications are less 'general', but these meters at
the outset were mostly for labs, not for servicing. Work in
instrumentation: sensors, transducers etc., can use uV precision
foremost, accuracy preferred. A diff DVM is ideal but not necessary,
a sensitive DMM is a cheaper substitute.
http://www.sensorland.com/HowPage078.html
At 12:09 PM 8/12/2013, J. Forster wrote:
My main interest is as a 'sanity check'. I use DMMs for engineering purposes.
Shields up:
IMO, there are very few applications, other than fundamental physics
research, that really NEED 5,6,7, or 8+ digits.
I would no more use a multi digit DVM to do what can be done with a
differential voltmeter than use a carrier phase tracking GPS to measure
for a living room carpet.
Shields down.
-John
Sincerely,
Marv
Philadelphia, PA