HP 34401A AC calibration error.

JP
John Phillips
Fri, Mar 2, 2012 4:46 PM

Hi,
I have been getting error 706 whenever I calibrate ACI.
706 read: Cal value out of range
The specified calibration value (CALibration:VALue) is invalid for the
present function and range.
Calibration requires setting the function before entering calibration.
Menu F: 2:calibrate  and the meter responds with 1.000,000 AAC.
Supposedly you are to apply 1 amp ant 1000 hz and press enter.
No matter what values I use or even change the 1.000 to some other value I
always get error 706 and the scale factor does not change.
Is there a way to calibrate the AC current without changing the shunt
resistor.
The shunt resister is 0.1 ohms at 1% yet the meter is to deliver 0.1% ACI.
Any ideas?

--
John Phillips

Hi, I have been getting error 706 whenever I calibrate ACI. 706 read: Cal value out of range The specified calibration value (CALibration:VALue) is invalid for the present function and range. Calibration requires setting the function before entering calibration. Menu F: 2:calibrate and the meter responds with 1.000,000 AAC. Supposedly you are to apply 1 amp ant 1000 hz and press enter. No matter what values I use or even change the 1.000 to some other value I always get error 706 and the scale factor does not change. Is there a way to calibrate the AC current without changing the shunt resistor. The shunt resister is 0.1 ohms at 1% yet the meter is to deliver 0.1% ACI. Any ideas? -- John Phillips