Fellow Voltage Fans,
To review - I would like to have a lab standard here that I can trust to
10 ppm or better to use to calibrate my expanding collection of
voltmeters. I have proposed to built a portable standard that I will
ship to participating volt nuts on this list who will record their
readings of the standard and share those readings with the list.
I have designed a Traveling Standard that incorporates two references
(an LM199AH and a MAX6350E) along with an LM34 to measure the
temperature of the MAX6350 (the LM34 is glued to the top of the MAX6350.
The schematic is attached to this page:
Following suggestions on this list (thanks!) I will use an external
"wall wart" supply for the input 24 Volts. This reduces any shift
caused by the line regulation of the three terminal regulator.
My design notes are included on the above page.
I have constructed a bread board of the circuit on a plugboard. I will
move the circuit to a printed circuit version after some testing.
Some preliminary readings from the breadboard of the LM199AH part are
noted on this page:
Note that these readings are taken with an HP3455A which has not been
calibrated since I bought it new. I would expect the accuracy to be no
better than .01% and may be much worse.
My plan is to mount the printed circuit board version in a short length
of 2" PVC pipe with connectors for the three outputs.
After the printed circuit board version is aged some it will be ready to
ship.
Are there any more on the list that would be willing to make a
measurement of the Traveling Standard?
I will collect the readings that we collectively get on the TS and make
them available to the list. I anticipate that with some experience we
can all benefit from having this TS well aged and with a well known
output voltage.
I will handle the shipping - all I ask is that you take careful
measurements and provide them to me, then return the unit.
Thanks!
Mhm,
im missing a 100nF at the output of the 7815.
and a 100 nF directly at the heater input of the LM399.
Also at the MAX6350 a 100nF is missing.
It is essential that these capacitors are located directly
at the pins of the respective ICs.
I have seen a 3,3V Voltage Regulator shifting output voltage by
several 100mV with a 3V/m electric field strength.
(which is quite low compared with a wireles phone of 30V/m or a mobile with
up to 200V/m).
And there was already a 10uF Tantal at the output.
The voltage shift could be elimininated with a 100nF directly
at input and output.
With best regards
Andreas
Hi Bob,
I have been monitoring your thread since it got started. Personally, you would
be better off spending your money on a used Fluke 731B off of eBay. You would be
way ahead of the game compared to where you are at with your current concept.
First, the biggest issue is not having your circuit set to a "cardinal" voltage
point; i.e., 10 volts. By having some other level the measuring party would have
to ratio against his standard to report a value. That adds an unnecessary level
of complexity and increases the error limits in reporting the uncertainty.
Second, you need to provide separate power supply regulators for any heating
system and the reference voltage circuitry.
Third, your choice of a power supply regulator is very poor. You should consider
something more exotic like the Linear Technology LT-3080 or the LT-3083. In
particular, pay attention to the noise reduction considerations in the
specification sheet. Preferably a two stage affair using a LT-3083 (handles 3
amps) as a pre regulator set to 18 volts and then the separate LT-3080 (handles 1
amp) devices to their respective circuitry.
Fourth, you need to pay particular attention to ground loops and your ground
plane layout. Think along the lines of a star pattern, i.e., single point of
return.
Fifth, to do it right you should provide four (4) separate 10 volt references
with their own outputs in the package. This would provide for independent
analysis of the "bank mean" and allows for determining if any individual
reference shifts outside of a nominal "bank mean."
Sixth, in designing the 10 volt reference, the output loading should provide the
highest possible isolation as to affecting the actual reference cell. The output
circuit should provide a minimum of 10 milliamps without a any change in the
output level.
Seventh, if any adjustable elements are used in the design, they should be of the
smallest possible size as to resistance value. This will help to minimize
temperature affects. A "cermet" potentiometer type should be used with values
under 50 Ohms. For example a 89pr10 (or 20 etc.).
Eighth, if it is going to be powered from the mains voltage, then a faraday
shielded transformer should be used. Proper EMI & RFI shielding is also needed.
Just food for thought and, by the way, having done all of the above, you just
constructed a 731 if you just do one reference.
Bill....WB6BNQ
Bob Smither wrote:
Fellow Voltage Fans,
To review - I would like to have a lab standard here that I can trust to
10 ppm or better to use to calibrate my expanding collection of
voltmeters. I have proposed to built a portable standard that I will
ship to participating volt nuts on this list who will record their
readings of the standard and share those readings with the list.
I have designed a Traveling Standard that incorporates two references
(an LM199AH and a MAX6350E) along with an LM34 to measure the
temperature of the MAX6350 (the LM34 is glued to the top of the MAX6350.
The schematic is attached to this page:
http://www.c-c-i.com/TS
Following suggestions on this list (thanks!) I will use an external
"wall wart" supply for the input 24 Volts. This reduces any shift
caused by the line regulation of the three terminal regulator.
My design notes are included on the above page.
I have constructed a bread board of the circuit on a plugboard. I will
move the circuit to a printed circuit version after some testing.
Some preliminary readings from the breadboard of the LM199AH part are
noted on this page:
http://c-c-i.com/node/121
Note that these readings are taken with an HP3455A which has not been
calibrated since I bought it new. I would expect the accuracy to be no
better than .01% and may be much worse.
My plan is to mount the printed circuit board version in a short length
of 2" PVC pipe with connectors for the three outputs.
After the printed circuit board version is aged some it will be ready to
ship.
Are there any more on the list that would be willing to make a
measurement of the Traveling Standard?
I will collect the readings that we collectively get on the TS and make
them available to the list. I anticipate that with some experience we
can all benefit from having this TS well aged and with a well known
output voltage.
I will handle the shipping - all I ask is that you take careful
measurements and provide them to me, then return the unit.
Thanks!
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Bob,
I am new to this but would very much like to participate. I have several
meters and would like the opportunity to make measurements with each.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Bob Smither
Sent: Monday, August 29, 2011 10:52 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Traveling Standards
Fellow Voltage Fans,
To review - I would like to have a lab standard here that I can trust to
10 ppm or better to use to calibrate my expanding collection of
voltmeters. I have proposed to built a portable standard that I will
ship to participating volt nuts on this list who will record their
readings of the standard and share those readings with the list.
I have designed a Traveling Standard that incorporates two references
(an LM199AH and a MAX6350E) along with an LM34 to measure the
temperature of the MAX6350 (the LM34 is glued to the top of the MAX6350.
The schematic is attached to this page:
Following suggestions on this list (thanks!) I will use an external
"wall wart" supply for the input 24 Volts. This reduces any shift
caused by the line regulation of the three terminal regulator.
My design notes are included on the above page.
I have constructed a bread board of the circuit on a plugboard. I will
move the circuit to a printed circuit version after some testing.
Some preliminary readings from the breadboard of the LM199AH part are
noted on this page:
Note that these readings are taken with an HP3455A which has not been
calibrated since I bought it new. I would expect the accuracy to be no
better than .01% and may be much worse.
My plan is to mount the printed circuit board version in a short length
of 2" PVC pipe with connectors for the three outputs.
After the printed circuit board version is aged some it will be ready to
ship.
Are there any more on the list that would be willing to make a
measurement of the Traveling Standard?
I will collect the readings that we collectively get on the TS and make
them available to the list. I anticipate that with some experience we
can all benefit from having this TS well aged and with a well known
output voltage.
I will handle the shipping - all I ask is that you take careful
measurements and provide them to me, then return the unit.
Thanks!
WB6BNQ wrote:
Hi Bob,
I have been monitoring your thread since it got started. Personally, you would
be better off spending your money on a used Fluke 731B off of eBay. You would be
way ahead of the game compared to where you are at with your current concept.
Hi Bill,
Thanks for taking the time to provide your very thoughtful feedback on
this project.
The last 731B I saw on EBay was ~$800.00. Unless I had it independently
calibrated (another expense) I would end up with another instrument in
my lab that I could not trust. Even if calibrated, after 30 days it
would have 10 ppm uncertainly and after a year 30 ppm.
First, the biggest issue is not having your circuit set to a "cardinal" voltage
point; i.e., 10 volts. By having some other level the measuring party would have
to ratio against his standard to report a value. That adds an unnecessary level
of complexity and increases the error limits in reporting the uncertainty.
My objective with this project is to obtain a stability of 10 ppm,
verified by having several independent readings made by other volt-nuts
members. If this objective can be realized, another advantage is the
unit can be passed around to help other list members.
Establishing a cardinal value such as 10 Volts would require an
adjustment. I am hoping to avoid that uncertainty. On a 10 volt scale
the ~7 Volt output of the LM199A provides 1.4 ppm resolution on a 6
digit VM - adequate for my objective of 10 ppm.
Second, you need to provide separate power supply regulators for any heating
system and the reference voltage circuitry.
Good point! I did overlook this. The heater of the LM199AH draws about
18 mA after it stabilizes and is fairly constant. It will certainly
change less than 5 mA over the expected temperature range of a lab. The
load regulation of the LM78L15 is (worst case) 1.5 mV / mA resulting in
an error at the LM199AH of 0.13 ppm. Although small, this can be
reduced further by powering the heater of the LM199AH directly from the
24 Volt supply and I will do that.
Third, your choice of a power supply regulator is very poor. You should consider
something more exotic like the Linear Technology LT-3080 or the LT-3083. In
particular, pay attention to the noise reduction considerations in the
specification sheet. Preferably a two stage affair using a LT-3083 (handles 3
amps) as a pre regulator set to 18 volts and then the separate LT-3080 (handles 1
amp) devices to their respective circuitry.
Although not shown directly on the posted schematic, the design does
provide two stage regulation. As detailed in my notes on the design
page ( http://www.c-c-i.com/ts ), there is a separate external 24 volt
supply. The specs on that supply are !% line regulation and .02%/C TC.
At the LM199AH, after secondary regulation from the LM78L15, these
errors translate to .06 ppm and .0011 ppm/C respectively.
The LT3080 is a great part. I want to keep the package as small as
possible, but I don't do surface mount. The through hole version of the
LT3080 is large compared to the LM78L15 (TO-92) that I am using.
The LM78L15 drifts at about 1 mV/C resulting in .018 ppm/C at the
LM199AH. If we assume that labs will be between 20 and 30 C this adds
.2 ppm.
I have an idea to substantially reduce this error source and will try it
when I get time.
Fourth, you need to pay particular attention to ground loops and your ground
plane layout. Think along the lines of a star pattern, i.e., single point of
return.
My instructions will be to measure only one output at a time. Each
output (from the LM199AH, the MAX6350E, and the LM34) has a separate
return terminal and the return current is brought back to the return pin
of each device, effectively removing the effect of other loads in the
ground wiring.
Fifth, to do it right you should provide four (4) separate 10 volt references
with their own outputs in the package. This would provide for independent
analysis of the "bank mean" and allows for determining if any individual
reference shifts outside of a nominal "bank mean."
Sixth, in designing the 10 volt reference, the output loading should provide the
highest possible isolation as to affecting the actual reference cell. The output
circuit should provide a minimum of 10 milliamps without a any change in the
output level.
The intent with this unit is that it will be measured using a voltmeter
or in a nulling arrangement. I am assuming that the input impedance of
the measuring instrument will be high. A 10Mohm load (.7 uA) on the
LM199AH introduces .1 ppm.
Seventh, if any adjustable elements are used in the design, they should be of the
smallest possible size as to resistance value. This will help to minimize
temperature affects. A "cermet" potentiometer type should be used with values
under 50 Ohms. For example a 89pr10 (or 20 etc.).
Eighth, if it is going to be powered from the mains voltage, then a faraday
shielded transformer should be used. Proper EMI & RFI shielding is also needed.
The external 24 Volt supply I picked has 240 mV p-p of noise. After the
LM78L15 this is reduced to 3 mV p-p or .018 ppm at the LM199AH.
Just food for thought and, by the way, having done all of the above, you just
constructed a 731 if you just do one reference.
To recap - my objective is a shared standard that can be inexpensively
shipped to other volt-nuts. After several of us have measured the unit
it will hopefully provide a reference that can be trusted to calibrate
other instruments, if only to the 10 ppm level.
Many thanks for your thoughts!
Bob Smither, PhD Circuit Concepts, Inc.
---=====
"We are over in Iraq trying to devise a new constitution for them.
Why don't we just give them ours? We're not using it any more."
--Jay Leno
---=====
Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744
Andreas Jahn wrote:
Mhm,
im missing a 100nF at the output of the 7815.
and a 100 nF directly at the heater input of the LM399.
Also at the MAX6350 a 100nF is missing.
It is essential that these capacitors are located directly
at the pins of the respective ICs.
Thanks for catching that!
I will have to adjust my packaging concept, but have added those parts.
See the updated schematic here:
I have seen a 3,3V Voltage Regulator shifting output voltage by
several 100mV with a 3V/m electric field strength.
I tried with a hand held walkie talkie - big effect. More notes to add
to the instructions that will go with the traveling standard. My cell
phone did not cause a measurable disturbance.
Bob Smither, Ph.D. Smither@c-c-i.com
---======
"The most potent weapon of the oppressor is the mind of the
oppressed." - Steven Biko
---======
Circuit Concepts, Inc. 281-331-2744
I have built the first traveling unit and it is under power and under test.
I have added some more notes and readings here:
The unit should be ready to ship to our first volunteer reader in a few
days.
---=
Bob Smither, Ph.D. Smither@c-c-i.com
Q: What is the square root of 4b^2?
A: To be or not to be.
---=
I finally found time to post some pictures of the TS:
--
Bob Smither, PhD Circuit Concepts, Inc.
---=====
Foreign aid might be defined as a transfer from poor people in
rich countries to rich people in poor countries.
--Douglas Casey (1992)
---=====
Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744(office)
Bob,
Is it not better to let it age a bit befor it starts its journey ? I have build several standards a few months ago. I have not used them since but then I meaured them during two weeks or so and yesterday I put them on again. They still are not perfect stable. The short time drift took a day or two to get smaller but the long time drift still is not perfect. ( i am talking about 40-60 uV for a 1V and a 2.5 V standard.)
Also test it mechanical. See if it reacts on shocks, temp differences ect.
Fred PA4TIM
Op 6 sep. 2011 om 19:04 heeft Bob Smither smither@c-c-i.com het volgende geschreven:
I finally found time to post some pictures of the TS:
--
Bob Smither, PhD Circuit Concepts, Inc.
---=====
Foreign aid might be defined as a transfer from poor people in
rich countries to rich people in poor countries.
--Douglas Casey (1992)
---=====
Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744(office)
<smither.vcf>
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Fred Schneider wrote:
Bob, Is it not better to let it age a bit befor it starts its journey
? I have build several standards a few months ago. I have not used
them since but then I meaured them during two weeks or so and
yesterday I put them on again. They still are not perfect stable. The
short time drift took a day or two to get smaller but the long time
drift still is not perfect. ( i am talking about 40-60 uV for a 1V
and a 2.5 V standard.)
Also test it mechanical. See if it reacts on shocks, temp differences
ect.
Thanks Fred. Both references (the LM119AH and the MAX6350) have several
hundred hours on them. I have left them on for all the time since I put
together the first bread board. When I was not working on them, they
have been powered. They will accumulate an additional hundred hours or
so before I actually ship them.
I will do some mechanical tests as well. The last thing I want to do is
"cook" the TS for a day or so at perhaps 100C and then see what has changed.
Thanks for the suggestions.
--
Bob Smither, PhD Circuit Concepts, Inc.
---=====
13th Amendment to the Constitution of the united States:
"Neither slavery nor involuntary servitude, except as a punishment for
crime whereof the party shall have been duly convicted, shall exist
within the United States, or any place subject to their jurisdiction.
Section 2.a of Proposed House Resolution 163, 108th Congress:
"It is the obligation of every citizen of the United States, and every
other person residing in the United States, who is between the ages of
18 and 26 to perform a period of national service ..."
---=====
Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744(office)