Discussion of precise voltage measurement

View all threads

732A and Prologix received

D
Don@True-Cal
Mon, Aug 25, 2014 3:02 PM

Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable  dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal.

The NI website had this chart that quantifies the Seebeck voltage best:
"When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients."

Junction µV/°C
Copper-Copper <0.3
Copper-Gold 0.5
Copper-Silver 0.5
Copper-Brass 3
Copper-Nickel 10
Copper-Lead-Tin Solder 1-3
Copper-Aluminum 5
Copper-Kovar 40
Copper-Copper Oxide >500

Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences.

Don Johnson

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Monday, August 25, 2014 5:37 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them.

Gesendet: Montag, 25. August 2014 um 06:33 Uhr
Von: "Orin Eman" orin.eman@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received

On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal truecalservices@gmail.com
wrote:

Randy & all,

You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.

11058A and 11174A are discontinued at Keysight.  However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs.  Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on these
as an alternative?

Orin.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal. The NI website had this chart that quantifies the Seebeck voltage best: "When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients." Junction µV/°C Copper-Copper <0.3 Copper-Gold 0.5 Copper-Silver 0.5 Copper-Brass 3 Copper-Nickel 10 Copper-Lead-Tin Solder 1-3 Copper-Aluminum 5 Copper-Kovar 40 Copper-Copper Oxide >500 Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences. Don Johnson -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de Sent: Monday, August 25, 2014 5:37 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 732A and Prologix received I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these. so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV). that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input): 1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them. > Gesendet: Montag, 25. August 2014 um 06:33 Uhr > Von: "Orin Eman" <orin.eman@gmail.com> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] 732A and Prologix received > > On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com> > wrote: > > > Randy & all, > > > > You have correctly concluded that some (maybe not all) of your measurement > > problem is thermal EMF being added or subtracted in series within your > > measurement interconnect. This thermal EMF is generated at the junction of > > dissimilar metals when accompanied with thermal gradients between the test > > lead and device terminals. You have to eliminate both the dissimilarity of > > the metal junctions as well as minimize the thermal differences. The > > terminals of the 3458A as well as the 732A are Beryllium Copper so you want > > to use the same test lead terminals. Forget the typical Tin plated lugs or > > even Gold plated as both are not Beryllium Copper and constitute dissimilar > > metals. The best solution (as usually the most expensive) is to use a set > > of > > Fluke 5440A-7005 (48") cables. I also have just as good results using the > > much more flexible Pomona 11174A (lugs end always stay connected to the > > 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable > > has the added Guard built in but be sure to also use a Guard lead with the > > Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY > > cables is usually not a good idea because the lead wire to terminal also > > constitutes just as critical of junction. The above cables use Tellurium > > Copper wire which is usually hard to find and hard to crimp properly and > > NEVER solder. > > > > > 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade > to banana cables are available (5295-36 at Mouser et al) and claim that > they are designed to minimize thermal EMFs. Datasheet is here: > http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these > as an alternative? > > Orin. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
A
acbern@gmx.de
Mon, Aug 25, 2014 3:48 PM

well, your last point is the issue, how can you have a temperature difference within a few microns of material in said connections. theory is one thing, but in reality it does not happen due to the givens of the setup.
therefore in practice it is irrelevant if the wire is silver or gold plated or pure copper. otherwise the gold plated spades and tellurium copper posts from pomona and others would be nonsense. and other than the mysterious fluke wire I have never seen a tellurium-copper wire from any wire manufacturer.

Gesendet: Montag, 25. August 2014 um 17:02 Uhr
Von: "Don@True-Cal" truecalservices@gmail.com
An: "'Discussion of precise voltage measurement'" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received

Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable  dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal.

The NI website had this chart that quantifies the Seebeck voltage best:
"When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients."

Junction µV/°C
Copper-Copper <0.3
Copper-Gold 0.5
Copper-Silver 0.5
Copper-Brass 3
Copper-Nickel 10
Copper-Lead-Tin Solder 1-3
Copper-Aluminum 5
Copper-Kovar 40
Copper-Copper Oxide >500

Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences.

Don Johnson

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de
Sent: Monday, August 25, 2014 5:37 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them.

Gesendet: Montag, 25. August 2014 um 06:33 Uhr
Von: "Orin Eman" orin.eman@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received

On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal truecalservices@gmail.com
wrote:

Randy & all,

You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.

11058A and 11174A are discontinued at Keysight.  However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs.  Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on these
as an alternative?

Orin.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

well, your last point is the issue, how can you have a temperature difference within a few microns of material in said connections. theory is one thing, but in reality it does not happen due to the givens of the setup. therefore in practice it is irrelevant if the wire is silver or gold plated or pure copper. otherwise the gold plated spades and tellurium copper posts from pomona and others would be nonsense. and other than the mysterious fluke wire I have never seen a tellurium-copper wire from any wire manufacturer. > Gesendet: Montag, 25. August 2014 um 17:02 Uhr > Von: "Don@True-Cal" <truecalservices@gmail.com> > An: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] 732A and Prologix received > > Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire due, as you say, for the malleability to crimp well and flexibility. The point I was making is to use the same interconnect test lead material throughout as the DUT terminal posts. The 3458A and the 732A both use Beryllium Copper alloy making that type interconnect lug or plug the best choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses pure copper rather than an alloy terminal and for the same reason, minimal Seebeck voltage is realized with a pure copper interconnect. Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal. > > The NI website had this chart that quantifies the Seebeck voltage best: > "When two, dissimilar metals are joined a voltage is created. This voltage is known as the thermal electromotive force (EMF) or the Seebeck voltage. The Seebeck voltage is dependent on the temperature of the junction and the composition of the metals joined. The specific metal-to-metal junctions result in specific temperature coefficients (µV/°C), also known as Seebeck coefficients. The following table lists the most common metals and their respective Seebeck coefficients." > > Junction µV/°C > Copper-Copper <0.3 > Copper-Gold 0.5 > Copper-Silver 0.5 > Copper-Brass 3 > Copper-Nickel 10 > Copper-Lead-Tin Solder 1-3 > Copper-Aluminum 5 > Copper-Kovar 40 > Copper-Copper Oxide >500 > > Granted, Gold and Silver are the next best choice, and is certainly why they are satisfactory, but using either warrants a more critical temperature gradient issue. If your measurements were satisfactorily convincing, than you probably had no appreciable junction temperature differences. > > Don Johnson > > -----Original Message----- > From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of acbern@gmx.de > Sent: Monday, August 25, 2014 5:37 AM > To: Discussion of precise voltage measurement > Subject: Re: [volt-nuts] 732A and Prologix received > > I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these. > so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV). > that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input): > 1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to copper-tellurium post or 4b.soldered copper connector(34420) my consistent results over more than a year using them. > > > > > Gesendet: Montag, 25. August 2014 um 06:33 Uhr > > Von: "Orin Eman" <orin.eman@gmail.com> > > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > > Betreff: Re: [volt-nuts] 732A and Prologix received > > > > On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com> > > wrote: > > > > > Randy & all, > > > > > > You have correctly concluded that some (maybe not all) of your measurement > > > problem is thermal EMF being added or subtracted in series within your > > > measurement interconnect. This thermal EMF is generated at the junction of > > > dissimilar metals when accompanied with thermal gradients between the test > > > lead and device terminals. You have to eliminate both the dissimilarity of > > > the metal junctions as well as minimize the thermal differences. The > > > terminals of the 3458A as well as the 732A are Beryllium Copper so you want > > > to use the same test lead terminals. Forget the typical Tin plated lugs or > > > even Gold plated as both are not Beryllium Copper and constitute dissimilar > > > metals. The best solution (as usually the most expensive) is to use a set > > > of > > > Fluke 5440A-7005 (48") cables. I also have just as good results using the > > > much more flexible Pomona 11174A (lugs end always stay connected to the > > > 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable > > > has the added Guard built in but be sure to also use a Guard lead with the > > > Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY > > > cables is usually not a good idea because the lead wire to terminal also > > > constitutes just as critical of junction. The above cables use Tellurium > > > Copper wire which is usually hard to find and hard to crimp properly and > > > NEVER solder. > > > > > > > > > 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade > > to banana cables are available (5295-36 at Mouser et al) and claim that > > they are designed to minimize thermal EMFs. Datasheet is here: > > http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these > > as an alternative? > > > > Orin. > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
MS
Mike S
Mon, Aug 25, 2014 4:02 PM

On 8/25/2014 11:02 AM, Don@True-Cal wrote:

Silver or Gold plating on the terminal or wire will introduce the
undesirable  dissimilar metal properties, both at the plating
junction and at the plating metal to DUT terminal.

Why?

Any Seebeck effect is immediately offset in the opposite direction,
since both junctions are (under normal conditions) at essentially the
same temperature (e.g. there's a copper-gold thermocouple, the minimal
thermal resistance of a micron of gold on the contact(s), then a
gold-copper thermocouple). It seems to me that the improved consistency
of the contact outweighs any loss from the thermocouples.

A more typical contact would be copper-nickel plate-gold plate, but the
concept is the same. Unless there is heat flowing through the entire
assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they
simply cancel.

Even if connecting gold plated to nickel plated contacts, it works out
the same - a copper-nickel-gold-nickel-copper connection is completely
offset. It's when the offsetting thermocouples occur across a
temperature gradient that you have problems.

--
Mike

On 8/25/2014 11:02 AM, Don@True-Cal wrote: > Silver or Gold plating on the terminal or wire will introduce the > undesirable dissimilar metal properties, both at the plating > junction and at the plating metal to DUT terminal. Why? Any Seebeck effect is immediately offset in the opposite direction, since both junctions are (under normal conditions) at essentially the same temperature (e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a micron of gold on the contact(s), then a gold-copper thermocouple). It seems to me that the improved consistency of the contact outweighs any loss from the thermocouples. A more typical contact would be copper-nickel plate-gold plate, but the concept is the same. Unless there is heat flowing through the entire assembly so one thermocouple is warmer than the offsetting one (e.g. shortly after plugging in a banana plug warmed by body heat), they simply cancel. Even if connecting gold plated to nickel plated contacts, it works out the same - a copper-nickel-gold-nickel-copper connection is completely offset. It's when the offsetting thermocouples occur across a temperature gradient that you have problems. -- Mike
CB
Charles Black
Mon, Aug 25, 2014 4:50 PM

Hi Adrian,

Although I have not used the Pomona spade lugs or a nano-voltmeter, my
experience is consistent with yours otherwise.

Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information
to deduce that lowest input short voltage is going to be a copper wire
since the meter is set to zero volts during calibration using a heavy
gauge (14 to 16) copper wire to short the top four input terminals. This
is very convenient since it is so easy to duplicate in the field and
makes simple inexpensive test leads best for high precision
measurements. In order for the 3458A to make full accuracy measurements
(8 digit) NPLC must be set to 1000 (according to the User's Guide). If
you use a lower NPLC value there is a table in the User's Guide that can
be used to determine how accurate your measurements are going to be.

Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been
about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.

Calibration shunt                        -0.00021mv +/- 10
pV            Equilibration time 5 minutes. 14 gauge per Calibration
Manual.
"U" heavy wire                            -0.00021mv +/-
10pV              Equil. time 5 minutes. Used the through holes in the
Input banana posts only.
Copper wire                                -0.00019mv +/-
10pV              NAPA PVC covered automobile wire at same contact
points as CAL shunt
Standard Ground Plate              -0.00040mv +/- 10pV Equil. time 5
minutes. Gold plated ground plate from my Datron 4910
Copper wire                                -0.00019mv +/-
10Pv              Equil. time 2 seconds. Used the banana through holes.
STP 2 meter test lead                -0.00021mv +/- 10pV
Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation.
Banana plugs                              -0.00021mv +/-
30pV              Equil. time 20 minutes. My best "Perfect" gold plated
plugs with copper wire.

Charlie

On 8/25/2014 3:36 AM, acbern@gmx.de wrote:

I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.

Gesendet: Montag, 25. August 2014 um 06:33 Uhr
Von: "Orin Eman" orin.eman@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received

On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal truecalservices@gmail.com
wrote:

Randy & all,

You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.

11058A and 11174A are discontinued at Keysight.  However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs.  Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on these
as an alternative?

Orin.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Adrian, Although I have not used the Pomona spade lugs or a nano-voltmeter, my experience is consistent with yours otherwise. Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information to deduce that lowest input short voltage is going to be a copper wire since the meter is set to zero volts during calibration using a heavy gauge (14 to 16) copper wire to short the top four input terminals. This is very convenient since it is so easy to duplicate in the field and makes simple inexpensive test leads best for high precision measurements. In order for the 3458A to make full accuracy measurements (8 digit) NPLC must be set to 1000 (according to the User's Guide). If you use a lower NPLC value there is a table in the User's Guide that can be used to determine how accurate your measurements are going to be. Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been about a year since I clid my last CAL 0 so it was going to be interesting at least for me. The first shunt was my test "U' shaped shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair. Calibration shunt -0.00021mv +/- 10 pV Equilibration time 5 minutes. 14 gauge per Calibration Manual. "U" heavy wire -0.00021mv +/- 10pV Equil. time 5 minutes. Used the through holes in the Input banana posts only. Copper wire -0.00019mv +/- 10pV NAPA PVC covered automobile wire at same contact points as CAL shunt Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5 minutes. Gold plated ground plate from my Datron 4910 Copper wire -0.00019mv +/- 10Pv Equil. time 2 seconds. Used the banana through holes. STP 2 meter test lead -0.00021mv +/- 10pV Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation. Banana plugs -0.00021mv +/- 30pV Equil. time 20 minutes. My best "Perfect" gold plated plugs with copper wire. Charlie On 8/25/2014 3:36 AM, acbern@gmx.de wrote: > I have used the pomona spades, mainly to interface the low emf pomona banana cables to binding posts. I have stopped this, reasons being, they are large and worse, that the pomona spring loaded insulation tube that covers the banana plug conductor uses such a strong spring that slowly the plug works its way out of the spade. this btw also happend to me when I used the pomona low emf binding posts together with the pomona low emf banana cables. overall I m not happy with these. > so, due to lack of options, I changed to self-made twisted shielded pair of high grade teflon/kapton silver plated copper cable with gold plated copper spades (crimped). I use them not only with the 3458a but also with nanovolt meters. these have higher resolution and accuracy in low level measurements than the 3458a. emf voltages were never an issue with these cables if properly used. I have posted some results doing 34420a stabilty measurements on the pmel forum, and the results are convincing (purpose was actually not to test the cables but the stability of the 34420a, but the emf issue is a part of this of course. we use the 34420a to do low voltage precision measurements on thermal converters where the full scale signal sometimes is 1mV). > that btw also relates to don's statements below, I do not concurr with his comments about copper telurium as cable and spade material and so on. this material, as stated here many times, is used because it is machinable, for copper spades one would not use it. the 34420a factory cable uses copper cable and copper spades, not telurium-copper. if there was a problem, it would be worse with the 34420a than with the 3458a because of its low level ranges. and again, I have not seen any problems in a chain of (output to input): > 1.copper-tellurium post from e.g. 8 digit calibrator > 2.crimped copper spade, gold plated > 3.silver plated tsp copper cable > 4a.crimped copper spade to copper-tellurium post or > 4b.soldered copper connector(34420) > my consistent results over more than a year using them. > > > >> Gesendet: Montag, 25. August 2014 um 06:33 Uhr >> Von: "Orin Eman" <orin.eman@gmail.com> >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> Betreff: Re: [volt-nuts] 732A and Prologix received >> >> On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservices@gmail.com> >> wrote: >> >>> Randy & all, >>> >>> You have correctly concluded that some (maybe not all) of your measurement >>> problem is thermal EMF being added or subtracted in series within your >>> measurement interconnect. This thermal EMF is generated at the junction of >>> dissimilar metals when accompanied with thermal gradients between the test >>> lead and device terminals. You have to eliminate both the dissimilarity of >>> the metal junctions as well as minimize the thermal differences. The >>> terminals of the 3458A as well as the 732A are Beryllium Copper so you want >>> to use the same test lead terminals. Forget the typical Tin plated lugs or >>> even Gold plated as both are not Beryllium Copper and constitute dissimilar >>> metals. The best solution (as usually the most expensive) is to use a set >>> of >>> Fluke 5440A-7005 (48") cables. I also have just as good results using the >>> much more flexible Pomona 11174A (lugs end always stay connected to the >>> 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable >>> has the added Guard built in but be sure to also use a Guard lead with the >>> Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY >>> cables is usually not a good idea because the lead wire to terminal also >>> constitutes just as critical of junction. The above cables use Tellurium >>> Copper wire which is usually hard to find and hard to crimp properly and >>> NEVER solder. >>> >> >> 11058A and 11174A are discontinued at Keysight. However, Pomona 5295 spade >> to banana cables are available (5295-36 at Mouser et al) and claim that >> they are designed to minimize thermal EMFs. Datasheet is here: >> http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on these >> as an alternative? >> >> Orin. >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > >
DM
Dave M
Mon, Aug 25, 2014 6:38 PM

Well, you sort of answered your own question.  The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple.  Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done.
Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer.  Leaving the reference voltage setting alone, a DC voltage is
input into the unit, and the DC source is adjusted for a null on the
galvanometer.  At that point, the AC voltage source is equal to that of the
DC voltage source.

Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard.  They aren't cheap, and you have to
have a converter for each range of voltages that you need to measure.  The
thermal converters used with this type of transfer standard isn't great (50
MHz or so typical), but their accuracy far surpasses that of the thermistor
type sensors.

There are other brands and models of thermal transfer standards, but I have
a Fluke model 540 and a few thermal converters.  That's why I referred you
to the manual for it.

Cheers,
Dave M

pa4tim@gmail.com wrote:

Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two AC+DC
TRMS 7,5 digit meters but the last calibration was 2 years ago.

My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for 1V,
10V and 100V (and use several resistors/heaters.)

Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that)  I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.

Fred, pa4tim

Well, you sort of answered your own question. The equipment is called a Thermal Transfer Standard, but instead of thermistors, it uses a thermocouple. Look at the manual for the Fluke 540B (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done. Basically, the AC source is input into the transfer standard, and the standard's internal reference voltage is adjusted for a null on the galvanometer. Leaving the reference voltage setting alone, a DC voltage is input into the unit, and the DC source is adjusted for a null on the galvanometer. At that point, the AC voltage source is equal to that of the DC voltage source. Ther are thermocouple-type thermal converters used for RF voltage measurements with the transfer standard. They aren't cheap, and you have to have a converter for each range of voltages that you need to measure. The thermal converters used with this type of transfer standard isn't great (50 MHz or so typical), but their accuracy far surpasses that of the thermistor type sensors. There are other brands and models of thermal transfer standards, but I have a Fluke model 540 and a few thermal converters. That's why I referred you to the manual for it. Cheers, Dave M pa4tim@gmail.com wrote: > Is there a way to link an AC voltage to a DC source for compare. I > can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) > against standardcells. But for AC I can not do that. I have two AC+DC > TRMS 7,5 digit meters but the last calibration was 2 years ago. > > My idea is in theory simple. It is based on the thermal converters > used in RF powermeters. Two resistors, two high resolution > temperature meters. AC on the first en DC on the second. If both are > the same temperature the AC voltage is the same as the DC voltage. > But I'm sure some people here have done this in the past. I would > like to use it for 50 to 100 kHz (or less) and something like for 1V, > 10V and 100V (and use several resistors/heaters.) > > Or mabey there is an other way to convert AC (for RF it can be done > with lightbubs but I never tryed that) I do not mind if it is slow > etc, I like this sort of experiments. You can learn a lot from it. > > Fred, pa4tim
A
acbern@gmx.de
Mon, Aug 25, 2014 7:05 PM

fred,
generally you raise a good point, I had the same issue of calibrating an ac voltage to a high level of accuracy. you need this e.g. to validate the self.cal of a 3458a or other precison stuff like the 8506a0.

what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need and a second one on range above. build a set of resistor range extenders (rf type with appropriate connectors and housings) to expand the range to where you need to be max. get one of the thermal converter calibrated (the higher one usually, and you need to havr  good cal lab, should be <10ppm accuracy) and use it to calibrate the rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for thermal converters! at higher frequencies, due to reflections and stray capacitance/inductance influences, the accuracy decreases. the resistor range extenders though, if build up correctly, only have a few ppm impact (there is a paper from nist on that, but this is only typical). you can calibrate all converters to the one you got externally calibrated. do some research in the web, when you do the calibration, you need to determine the so-called constant N. then do an ac, dc+, ac, dc-, ac measurement between the the two and establish the deviation, also establish the error propagation. the end result will be a set of highly precise (low inaccuracies9 thermal converters good enough to calibrate a 3458a an better devices. if you want to spend the money, you could also buy a set of converters/range resistors (with/without a 540), that typically is a few k altogether, while a single device sometimes is available for below 100 bucks. you need to have a stable 7.5 digit nanovoltmeter though for the measurements of the tvcs (34420a or 2182 typically ) and precision (stable) dc and ac sources. but in the end, all you need is a single calibrated thermal converter.

adrian

Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration

Well, you sort of answered your own question.  The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple.  Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done.
Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer.  Leaving the reference voltage setting alone, a DC voltage is
input into the unit, and the DC source is adjusted for a null on the
galvanometer.  At that point, the AC voltage source is equal to that of the
DC voltage source.

Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard.  They aren't cheap, and you have to
have a converter for each range of voltages that you need to measure.  The
thermal converters used with this type of transfer standard isn't great (50
MHz or so typical), but their accuracy far surpasses that of the thermistor
type sensors.

There are other brands and models of thermal transfer standards, but I have
a Fluke model 540 and a few thermal converters.  That's why I referred you
to the manual for it.

Cheers,
Dave M

pa4tim@gmail.com wrote:

Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two AC+DC
TRMS 7,5 digit meters but the last calibration was 2 years ago.

My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for 1V,
10V and 100V (and use several resistors/heaters.)

Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that)  I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.

Fred, pa4tim


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

fred, generally you raise a good point, I had the same issue of calibrating an ac voltage to a high level of accuracy. you need this e.g. to validate the self.cal of a 3458a or other precison stuff like the 8506a0. what i would recommend to do if you want to keep costs down is: in a nutshell, get a thermal converter in the lowest range you need and a second one on range above. build a set of resistor range extenders (rf type with appropriate connectors and housings) to expand the range to where you need to be max. get one of the thermal converter calibrated (the higher one usually, and you need to havr good cal lab, should be <10ppm accuracy) and use it to calibrate the rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for thermal converters! at higher frequencies, due to reflections and stray capacitance/inductance influences, the accuracy decreases. the resistor range extenders though, if build up correctly, only have a few ppm impact (there is a paper from nist on that, but this is only typical). you can calibrate all converters to the one you got externally calibrated. do some research in the web, when you do the calibration, you need to determine the so-called constant N. then do an ac, dc+, ac, dc-, ac measurement between the the two and establish the deviation, also establish the error propagation. the end result will be a set of highly precise (low inaccuracies9 thermal converters good enough to calibrate a 3458a an better devices. if you want to spend the money, you could also buy a set of converters/range resistors (with/without a 540), that typically is a few k altogether, while a single device sometimes is available for below 100 bucks. you need to have a stable 7.5 digit nanovoltmeter though for the measurements of the tvcs (34420a or 2182 typically ) and precision (stable) dc and ac sources. but in the end, all you need is a single calibrated thermal converter. adrian > Gesendet: Montag, 25. August 2014 um 18:38 Uhr > Von: "Dave M" <dgminala@mediacombb.net> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] AC calibration > > Well, you sort of answered your own question. The equipment is called a > Thermal Transfer Standard, but instead of thermistors, it uses a > thermocouple. Look at the manual for the Fluke 540B > (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done. > Basically, the AC source is input into the transfer standard, and the > standard's internal reference voltage is adjusted for a null on the > galvanometer. Leaving the reference voltage setting alone, a DC voltage is > input into the unit, and the DC source is adjusted for a null on the > galvanometer. At that point, the AC voltage source is equal to that of the > DC voltage source. > > Ther are thermocouple-type thermal converters used for RF voltage > measurements with the transfer standard. They aren't cheap, and you have to > have a converter for each range of voltages that you need to measure. The > thermal converters used with this type of transfer standard isn't great (50 > MHz or so typical), but their accuracy far surpasses that of the thermistor > type sensors. > > There are other brands and models of thermal transfer standards, but I have > a Fluke model 540 and a few thermal converters. That's why I referred you > to the manual for it. > > Cheers, > Dave M > > > pa4tim@gmail.com wrote: > > Is there a way to link an AC voltage to a DC source for compare. I > > can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) > > against standardcells. But for AC I can not do that. I have two AC+DC > > TRMS 7,5 digit meters but the last calibration was 2 years ago. > > > > My idea is in theory simple. It is based on the thermal converters > > used in RF powermeters. Two resistors, two high resolution > > temperature meters. AC on the first en DC on the second. If both are > > the same temperature the AC voltage is the same as the DC voltage. > > But I'm sure some people here have done this in the past. I would > > like to use it for 50 to 100 kHz (or less) and something like for 1V, > > 10V and 100V (and use several resistors/heaters.) > > > > Or mabey there is an other way to convert AC (for RF it can be done > > with lightbubs but I never tryed that) I do not mind if it is slow > > etc, I like this sort of experiments. You can learn a lot from it. > > > > Fred, pa4tim > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
CH
Chuck Harris
Mon, Aug 25, 2014 7:12 PM

Look up the phrase "AC thermal transfer standard".

Using a heater/thermocouple element in a vacuum is the tried
and true way of linking an AC voltage to a DC reference.

-Chuck Harris

OBTW, trimming your quoted posts is considered friendly.

pa4tim@gmail.com wrote:

Is there a way to link an AC voltage to a DC source for compare. I can check my
calibrators (like a Fluke 332, 760 , 731 and a Philips)  against standardcells.
But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last
calibration was 2 years ago.

My idea is in theory simple. It is based on the thermal converters used in RF
powermeters. Two resistors, two high resolution temperature meters. AC on the
first en DC on the second. If both are the same temperature the AC voltage is the
same as the DC voltage. But I'm sure some people here have done this in the past.
I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V
and 100V (and use several resistors/heaters.)

Or mabey there is an other way to convert AC (for RF it can be done with lightbubs
but I never tryed that)  I do not mind if it is slow etc, I like this sort of
experiments. You can learn a lot from it.

Fred, pa4tim

Look up the phrase "AC thermal transfer standard". Using a heater/thermocouple element in a vacuum is the tried and true way of linking an AC voltage to a DC reference. -Chuck Harris OBTW, trimming your quoted posts is considered friendly. pa4tim@gmail.com wrote: > Is there a way to link an AC voltage to a DC source for compare. I can check my > calibrators (like a Fluke 332, 760 , 731 and a Philips) against standardcells. > But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last > calibration was 2 years ago. > > > My idea is in theory simple. It is based on the thermal converters used in RF > powermeters. Two resistors, two high resolution temperature meters. AC on the > first en DC on the second. If both are the same temperature the AC voltage is the > same as the DC voltage. But I'm sure some people here have done this in the past. > I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V > and 100V (and use several resistors/heaters.) > > > Or mabey there is an other way to convert AC (for RF it can be done with lightbubs > but I never tryed that) I do not mind if it is slow etc, I like this sort of > experiments. You can learn a lot from it. > > > > > Fred, pa4tim
CB
Charles Black
Mon, Aug 25, 2014 7:28 PM

I just realized that all my measurement tolerances are  nV not pV. Sorry
about not checking units first.

Charlie

On 8/25/2014 9:50 AM, Charles Black wrote:

Hi Adrian,

Although I have not used the Pomona spade lugs or a nano-voltmeter, my
experience is consistent with yours otherwise.

Anyone who has calibrated (CAL 0 anyway) a 3458A has enough
information to deduce that lowest input short voltage is going to be a
copper wire since the meter is set to zero volts during calibration
using a heavy gauge (14 to 16) copper wire to short the top four input
terminals. This is very convenient since it is so easy to duplicate in
the field and makes simple inexpensive test leads best for high
precision measurements. In order for the 3458A to make full accuracy
measurements (8 digit) NPLC must be set to 1000 (according to the
User's Guide). If you use a lower NPLC value there is a table in the
User's Guide that can be used to determine how accurate your
measurements are going to be.

Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has
been about a year since I clid my last CAL 0 so it was going to be
interesting at least for me. The first shunt was my test "U' shaped
shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.

Calibration shunt                        -0.00021mv +/- 10
pV            Equilibration time 5 minutes. 14 gauge per Calibration
Manual.
"U" heavy wire                            -0.00021mv +/-
10pV              Equil. time 5 minutes. Used the through holes in the
Input banana posts only.
Copper wire                                -0.00019mv +/-
10pV              NAPA PVC covered automobile wire at same contact
points as CAL shunt
Standard Ground Plate              -0.00040mv +/- 10pV Equil. time 5
minutes. Gold plated ground plate from my Datron 4910
Copper wire                                -0.00019mv +/-
10Pv              Equil. time 2 seconds. Used the banana through holes.
STP 2 meter test lead                -0.00021mv +/- 10pV
Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation.
Banana plugs                              -0.00021mv +/-
30pV              Equil. time 20 minutes. My best "Perfect" gold
plated plugs with copper wire.

Charlie

On 8/25/2014 3:36 AM, acbern@gmx.de wrote:

I have used the pomona spades, mainly to interface the low emf pomona
banana cables to binding posts. I have stopped this, reasons being,
they are large and worse, that the pomona spring loaded insulation
tube that covers the banana plug conductor uses such a strong spring
that slowly the plug works its way out of the spade. this btw also
happend to me when I used the pomona low emf binding posts together
with the pomona low emf banana cables. overall I m not happy with these.
so, due to lack of options, I changed to self-made twisted shielded
pair of high grade teflon/kapton silver plated copper cable with gold
plated copper spades (crimped). I use them not only with the 3458a
but also with nanovolt meters. these have higher resolution and
accuracy in low level measurements than the 3458a. emf voltages were
never an issue with these cables if properly used. I have posted some
results doing 34420a stabilty measurements on the pmel forum, and the
results are convincing (purpose was actually not to test the cables
but the stability of the 34420a, but the emf issue is a part of this
of course. we use the 34420a to do low voltage precision measurements
on thermal converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr
with his comments about copper telurium as cable and spade material
and so on. this material, as stated here many times, is used because
it is machinable, for copper spades one would not use it. the 34420a
factory cable uses copper cable and copper spades, not
telurium-copper. if there was a problem, it would be worse with the
34420a than with the 3458a because of its low level ranges. and
again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.

Gesendet: Montag, 25. August 2014 um 06:33 Uhr
Von: "Orin Eman" orin.eman@gmail.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] 732A and Prologix received

On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal
truecalservices@gmail.com
wrote:

Randy & all,

You have correctly concluded that some (maybe not all) of your
measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the
junction of
dissimilar metals when accompanied with thermal gradients between
the test
lead and device terminals. You have to eliminate both the
dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so
you want
to use the same test lead terminals. Forget the typical Tin plated
lugs or
even Gold plated as both are not Beryllium Copper and constitute
dissimilar
metals. The best solution (as usually the most expensive) is to use
a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results
using the
much more flexible Pomona 11174A (lugs end always stay connected to
the
732A) or 11058A with more convenient shielded banana plugs. The
Fluke cable
has the added Guard built in but be sure to also use a Guard lead
with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal
also
constitutes just as critical of junction. The above cables use
Tellurium
Copper wire which is usually hard to find and hard to crimp
properly and
NEVER solder.

11058A and 11174A are discontinued at Keysight.  However, Pomona
5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs.  Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on
these
as an alternative?

Orin.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I just realized that all my measurement tolerances are nV not pV. Sorry about not checking units first. Charlie On 8/25/2014 9:50 AM, Charles Black wrote: > Hi Adrian, > > Although I have not used the Pomona spade lugs or a nano-voltmeter, my > experience is consistent with yours otherwise. > > Anyone who has calibrated (CAL 0 anyway) a 3458A has enough > information to deduce that lowest input short voltage is going to be a > copper wire since the meter is set to zero volts during calibration > using a heavy gauge (14 to 16) copper wire to short the top four input > terminals. This is very convenient since it is so easy to duplicate in > the field and makes simple inexpensive test leads best for high > precision measurements. In order for the 3458A to make full accuracy > measurements (8 digit) NPLC must be set to 1000 (according to the > User's Guide). If you use a lower NPLC value there is a table in the > User's Guide that can be used to determine how accurate your > measurements are going to be. > > Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has > been about a year since I clid my last CAL 0 so it was going to be > interesting at least for me. The first shunt was my test "U' shaped > shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair. > > Calibration shunt -0.00021mv +/- 10 > pV Equilibration time 5 minutes. 14 gauge per Calibration > Manual. > "U" heavy wire -0.00021mv +/- > 10pV Equil. time 5 minutes. Used the through holes in the > Input banana posts only. > Copper wire -0.00019mv +/- > 10pV NAPA PVC covered automobile wire at same contact > points as CAL shunt > Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5 > minutes. Gold plated ground plate from my Datron 4910 > Copper wire -0.00019mv +/- > 10Pv Equil. time 2 seconds. Used the banana through holes. > STP 2 meter test lead -0.00021mv +/- 10pV > Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation. > Banana plugs -0.00021mv +/- > 30pV Equil. time 20 minutes. My best "Perfect" gold > plated plugs with copper wire. > > Charlie > > > On 8/25/2014 3:36 AM, acbern@gmx.de wrote: >> I have used the pomona spades, mainly to interface the low emf pomona >> banana cables to binding posts. I have stopped this, reasons being, >> they are large and worse, that the pomona spring loaded insulation >> tube that covers the banana plug conductor uses such a strong spring >> that slowly the plug works its way out of the spade. this btw also >> happend to me when I used the pomona low emf binding posts together >> with the pomona low emf banana cables. overall I m not happy with these. >> so, due to lack of options, I changed to self-made twisted shielded >> pair of high grade teflon/kapton silver plated copper cable with gold >> plated copper spades (crimped). I use them not only with the 3458a >> but also with nanovolt meters. these have higher resolution and >> accuracy in low level measurements than the 3458a. emf voltages were >> never an issue with these cables if properly used. I have posted some >> results doing 34420a stabilty measurements on the pmel forum, and the >> results are convincing (purpose was actually not to test the cables >> but the stability of the 34420a, but the emf issue is a part of this >> of course. we use the 34420a to do low voltage precision measurements >> on thermal converters where the full scale signal sometimes is 1mV). >> that btw also relates to don's statements below, I do not concurr >> with his comments about copper telurium as cable and spade material >> and so on. this material, as stated here many times, is used because >> it is machinable, for copper spades one would not use it. the 34420a >> factory cable uses copper cable and copper spades, not >> telurium-copper. if there was a problem, it would be worse with the >> 34420a than with the 3458a because of its low level ranges. and >> again, I have not seen any problems in a chain of (output to input): >> 1.copper-tellurium post from e.g. 8 digit calibrator >> 2.crimped copper spade, gold plated >> 3.silver plated tsp copper cable >> 4a.crimped copper spade to copper-tellurium post or >> 4b.soldered copper connector(34420) >> my consistent results over more than a year using them. >> >> >> >>> Gesendet: Montag, 25. August 2014 um 06:33 Uhr >>> Von: "Orin Eman" <orin.eman@gmail.com> >>> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >>> Betreff: Re: [volt-nuts] 732A and Prologix received >>> >>> On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal >>> <truecalservices@gmail.com> >>> wrote: >>> >>>> Randy & all, >>>> >>>> You have correctly concluded that some (maybe not all) of your >>>> measurement >>>> problem is thermal EMF being added or subtracted in series within your >>>> measurement interconnect. This thermal EMF is generated at the >>>> junction of >>>> dissimilar metals when accompanied with thermal gradients between >>>> the test >>>> lead and device terminals. You have to eliminate both the >>>> dissimilarity of >>>> the metal junctions as well as minimize the thermal differences. The >>>> terminals of the 3458A as well as the 732A are Beryllium Copper so >>>> you want >>>> to use the same test lead terminals. Forget the typical Tin plated >>>> lugs or >>>> even Gold plated as both are not Beryllium Copper and constitute >>>> dissimilar >>>> metals. The best solution (as usually the most expensive) is to use >>>> a set >>>> of >>>> Fluke 5440A-7005 (48") cables. I also have just as good results >>>> using the >>>> much more flexible Pomona 11174A (lugs end always stay connected to >>>> the >>>> 732A) or 11058A with more convenient shielded banana plugs. The >>>> Fluke cable >>>> has the added Guard built in but be sure to also use a Guard lead >>>> with the >>>> Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY >>>> cables is usually not a good idea because the lead wire to terminal >>>> also >>>> constitutes just as critical of junction. The above cables use >>>> Tellurium >>>> Copper wire which is usually hard to find and hard to crimp >>>> properly and >>>> NEVER solder. >>>> >>> >>> 11058A and 11174A are discontinued at Keysight. However, Pomona >>> 5295 spade >>> to banana cables are available (5295-36 at Mouser et al) and claim that >>> they are designed to minimize thermal EMFs. Datasheet is here: >>> http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf Any comments on >>> these >>> as an alternative? >>> >>> Orin. >>> _______________________________________________ >>> volt-nuts mailing list -- volt-nuts@febo.com >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>> and follow the instructions there. >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> >> >> > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > >
DM
Dave M
Mon, Aug 25, 2014 7:32 PM

Adrian,
Do you have a link or title for the NIST paper that you mentioned?

Dave M

acbern@gmx.de wrote:

fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.

what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be <10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.

adrian

Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: "Dave M" dgminala@mediacombb.net
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration

Well, you sort of answered your own question.  The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple.  Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer.  Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer.  At that point,
the AC voltage source is equal to that of the DC voltage source.

Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard.  They aren't cheap, and you
have to have a converter for each range of voltages that you need to
measure.  The thermal converters used with this type of transfer
standard isn't great (50 MHz or so typical), but their accuracy far
surpasses that of the thermistor type sensors.

There are other brands and models of thermal transfer standards, but
I have a Fluke model 540 and a few thermal converters.  That's why I
referred you to the manual for it.

Cheers,
Dave M

pa4tim@gmail.com wrote:

Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two
AC+DC TRMS 7,5 digit meters but the last calibration was 2 years
ago.

My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for
1V, 10V and 100V (and use several resistors/heaters.)

Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that)  I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.

Fred, pa4tim


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

When the people fear the government, there is tyranny. When the
government fears the people, there is liberty -- Thomas Jefferson

Dave M

Adrian, Do you have a link or title for the NIST paper that you mentioned? Dave M acbern@gmx.de wrote: > fred, > generally you raise a good point, I had the same issue of calibrating > an ac voltage to a high level of accuracy. you need this e.g. to > validate the self.cal of a 3458a or other precison stuff like the > 8506a0. > > what i would recommend to do if you want to keep costs down is: > in a nutshell, get a thermal converter in the lowest range you need > and a second one on range above. build a set of resistor range > extenders (rf type with appropriate connectors and housings) to > expand the range to where you need to be max. get one of the thermal > converter calibrated (the higher one usually, and you need to havr > good cal lab, should be <10ppm accuracy) and use it to calibrate the > rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for > thermal converters! at higher frequencies, due to reflections and > stray capacitance/inductance influences, the accuracy decreases. the > resistor range extenders though, if build up correctly, only have a > few ppm impact (there is a paper from nist on that, but this is only > typical). you can calibrate all converters to the one you got > externally calibrated. do some research in the web, when you do the > calibration, you need to determine the so-called constant N. then do > an ac, dc+, ac, dc-, ac measurement between the the two and establish > the deviation, also establish the error propagation. the end result > will be a set of highly precise (low inaccuracies9 thermal converters > good enough to calibrate a 3458a an better devices. if you want to > spend the money, you could also buy a set of converters/range > resistors (with/without a 540), that typically is a few k altogether, > while a single device sometimes is available for below 100 bucks. you > need to have a stable 7.5 digit nanovoltmeter though for the > measurements of the tvcs (34420a or 2182 typically ) and precision > (stable) dc and ac sources. but in the end, all you need is a single > calibrated thermal converter. > > adrian > > > >> Gesendet: Montag, 25. August 2014 um 18:38 Uhr >> Von: "Dave M" <dgminala@mediacombb.net> >> An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> >> Betreff: Re: [volt-nuts] AC calibration >> >> Well, you sort of answered your own question. The equipment is >> called a Thermal Transfer Standard, but instead of thermistors, it >> uses a thermocouple. Look at the manual for the Fluke 540B >> (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how >> it's done. Basically, the AC source is input into the transfer >> standard, and the standard's internal reference voltage is adjusted >> for a null on the galvanometer. Leaving the reference voltage >> setting alone, a DC voltage is input into the unit, and the DC >> source is adjusted for a null on the galvanometer. At that point, >> the AC voltage source is equal to that of the DC voltage source. >> >> Ther are thermocouple-type thermal converters used for RF voltage >> measurements with the transfer standard. They aren't cheap, and you >> have to have a converter for each range of voltages that you need to >> measure. The thermal converters used with this type of transfer >> standard isn't great (50 MHz or so typical), but their accuracy far >> surpasses that of the thermistor type sensors. >> >> There are other brands and models of thermal transfer standards, but >> I have a Fluke model 540 and a few thermal converters. That's why I >> referred you to the manual for it. >> >> Cheers, >> Dave M >> >> >> pa4tim@gmail.com wrote: >>> Is there a way to link an AC voltage to a DC source for compare. I >>> can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) >>> against standardcells. But for AC I can not do that. I have two >>> AC+DC TRMS 7,5 digit meters but the last calibration was 2 years >>> ago. >>> >>> My idea is in theory simple. It is based on the thermal converters >>> used in RF powermeters. Two resistors, two high resolution >>> temperature meters. AC on the first en DC on the second. If both are >>> the same temperature the AC voltage is the same as the DC voltage. >>> But I'm sure some people here have done this in the past. I would >>> like to use it for 50 to 100 kHz (or less) and something like for >>> 1V, 10V and 100V (and use several resistors/heaters.) >>> >>> Or mabey there is an other way to convert AC (for RF it can be done >>> with lightbubs but I never tryed that) I do not mind if it is slow >>> etc, I like this sort of experiments. You can learn a lot from it. >>> >>> Fred, pa4tim >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow >> the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. When the people fear the government, there is tyranny. When the government fears the people, there is liberty -- Thomas Jefferson Dave M
BG
Bill Gold
Mon, Aug 25, 2014 10:09 PM

There is one more thing that enters into this discussion and that is
"reversal errors" on the DC.  The complicates the transfer somewhat.  AC is
always going "+" and "-".  DC is in one direction so you have to then
reverse the voltage to the Thermal Transfer Standard and then take the
average of the two readings.  That is why there is a "Reversal" switch on
the 540B.  When you are using a fixed voltage High Frequency Thermal
Converter you need an external DC reversal switch in addition to other
equipment.  You also need an AC/DC transfer switch so that you don't have to
disconnect the AC source and then hook up the DC source manually.  See the
540B again.

All Thermal Transfer Standards have some "reversal" error.  This is
controlled by the internal construction of the unit and exactly where the
glass isolation bead is located on the heating element.  The thermocouple
converter used in the 540B is selected to have a very low reversal error,
but always will have some error.  The error is fixed so you can approximate
a DC measurement once you have characterized the particular converter.  I
can't remember now but I think there can be up to around .05% reversal
errors on some converters, while the ones selected for the 540B are under
.01%.

Read the FLUKE "Calibration: Philosophy in Practice" for further
information.

Bill

----- Original Message -----
From: "Dave M" dgminala@mediacombb.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, August 25, 2014 11:38 AM
Subject: Re: [volt-nuts] AC calibration

Well, you sort of answered your own question.  The equipment is called a
Thermal Transfer Standard, but instead of thermistors, it uses a
thermocouple.  Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's

done.

Basically, the AC source is input into the transfer standard, and the
standard's internal reference voltage is adjusted for a null on the
galvanometer.  Leaving the reference voltage setting alone, a DC voltage

is

input into the unit, and the DC source is adjusted for a null on the
galvanometer.  At that point, the AC voltage source is equal to that of

the

DC voltage source.

Ther are thermocouple-type thermal converters used for RF voltage
measurements with the transfer standard.  They aren't cheap, and you have

to

have a converter for each range of voltages that you need to measure.  The
thermal converters used with this type of transfer standard isn't great

(50

MHz or so typical), but their accuracy far surpasses that of the

thermistor

type sensors.

There are other brands and models of thermal transfer standards, but I

have

a Fluke model 540 and a few thermal converters.  That's why I referred you
to the manual for it.

Cheers,
Dave M

pa4tim@gmail.com wrote:

Is there a way to link an AC voltage to a DC source for compare. I
can check my calibrators (like a Fluke 332, 760 , 731 and a Philips)
against standardcells. But for AC I can not do that. I have two AC+DC
TRMS 7,5 digit meters but the last calibration was 2 years ago.

My idea is in theory simple. It is based on the thermal converters
used in RF powermeters. Two resistors, two high resolution
temperature meters. AC on the first en DC on the second. If both are
the same temperature the AC voltage is the same as the DC voltage.
But I'm sure some people here have done this in the past. I would
like to use it for 50 to 100 kHz (or less) and something like for 1V,
10V and 100V (and use several resistors/heaters.)

Or mabey there is an other way to convert AC (for RF it can be done
with lightbubs but I never tryed that)  I do not mind if it is slow
etc, I like this sort of experiments. You can learn a lot from it.

Fred, pa4tim


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.

There is one more thing that enters into this discussion and that is "reversal errors" on the DC. The complicates the transfer somewhat. AC is always going "+" and "-". DC is in one direction so you have to then reverse the voltage to the Thermal Transfer Standard and then take the average of the two readings. That is why there is a "Reversal" switch on the 540B. When you are using a fixed voltage High Frequency Thermal Converter you need an external DC reversal switch in addition to other equipment. You also need an AC/DC transfer switch so that you don't have to disconnect the AC source and then hook up the DC source manually. See the 540B again. All Thermal Transfer Standards have some "reversal" error. This is controlled by the internal construction of the unit and exactly where the glass isolation bead is located on the heating element. The thermocouple converter used in the 540B is selected to have a very low reversal error, but always will have some error. The error is fixed so you can approximate a DC measurement once you have characterized the particular converter. I can't remember now but I think there can be up to around .05% reversal errors on some converters, while the ones selected for the 540B are under .01%. Read the FLUKE "Calibration: Philosophy in Practice" for further information. Bill ----- Original Message ----- From: "Dave M" <dgminala@mediacombb.net> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Monday, August 25, 2014 11:38 AM Subject: Re: [volt-nuts] AC calibration > Well, you sort of answered your own question. The equipment is called a > Thermal Transfer Standard, but instead of thermistors, it uses a > thermocouple. Look at the manual for the Fluke 540B > (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done. > Basically, the AC source is input into the transfer standard, and the > standard's internal reference voltage is adjusted for a null on the > galvanometer. Leaving the reference voltage setting alone, a DC voltage is > input into the unit, and the DC source is adjusted for a null on the > galvanometer. At that point, the AC voltage source is equal to that of the > DC voltage source. > > Ther are thermocouple-type thermal converters used for RF voltage > measurements with the transfer standard. They aren't cheap, and you have to > have a converter for each range of voltages that you need to measure. The > thermal converters used with this type of transfer standard isn't great (50 > MHz or so typical), but their accuracy far surpasses that of the thermistor > type sensors. > > There are other brands and models of thermal transfer standards, but I have > a Fluke model 540 and a few thermal converters. That's why I referred you > to the manual for it. > > Cheers, > Dave M > > > pa4tim@gmail.com wrote: > > Is there a way to link an AC voltage to a DC source for compare. I > > can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) > > against standardcells. But for AC I can not do that. I have two AC+DC > > TRMS 7,5 digit meters but the last calibration was 2 years ago. > > > > My idea is in theory simple. It is based on the thermal converters > > used in RF powermeters. Two resistors, two high resolution > > temperature meters. AC on the first en DC on the second. If both are > > the same temperature the AC voltage is the same as the DC voltage. > > But I'm sure some people here have done this in the past. I would > > like to use it for 50 to 100 kHz (or less) and something like for 1V, > > 10V and 100V (and use several resistors/heaters.) > > > > Or mabey there is an other way to convert AC (for RF it can be done > > with lightbubs but I never tryed that) I do not mind if it is slow > > etc, I like this sort of experiments. You can learn a lot from it. > > > > Fred, pa4tim > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.