BG
Bruce Griffiths
Sat, Jul 4, 2009 5:40 AM
Warren
I agree, it's certainly possible that the external circuitry should have
a higher operating voltage than 5V.
A crystal current of 50uA to 1mA or so might perhaps be expected for a
typical 10MHz overtone crystal.
Unless some non standard crystal cut with exceptionally high ESR is used.
The oscillator transistor dc emitter current appears consistent with a
crystal current of around 1mA or so.
Can use Spice to check but there is no need for elaborate starting
techniques when simulating this circuit as it has plenty of excess gain
if the simulated Q isn't too large (50,000 or so is easy) (only ESR has
much effect on the crystal current).
Elaborate simulations looking for ppm changes when the active device
models aren't that good can be counterproductive.
Doing this with any confidence requires enormous trust in the
convergence of the numerical algorithms employed.
All flavours of Spice do have problems with accuracy and convergence
when the condition number of the solution matrix is large enough.
Often this is because the simulated circuit is unstable so an educated
guess as to what to change to preserve stability/convergence is required.
Occasionally Spice can even predict noise that is much larger than that
actually measured (it can have trouble when one has both positive and
negative feedback loops).
Bruce
WarrenS wrote:
Thus one cannot arbitrarily increase Vcc without considering the consequences ...
True, but with the AC voltages as low as they are, I do think it is safe to think that the crystal current should be higher.
I was not suggesting to arbitrarily rise the Vcc, I was asking any chance it is at the wrong voltage???
Even though 2.49 volts is plenty of Vcc to design an osc circuit to run on, It seems like a low, funny voltage to design for, on what I assume is an old product.
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:36 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
I agree with Warren. The VCC seems way too low. Interestingly, noone else seems to think so,
or did not notice it.
Bill....WB6BNQ
It was noticed, but it is consistent with the resistor values used in
the circuit.
The crystal current is determined by the dc emitter current of the
oscillator transistor.
Thus one cannot arbitrarily increase Vcc without considering the
consequences for the crystal current etc.
Bruce
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
Warren
I agree, it's certainly possible that the external circuitry should have
a higher operating voltage than 5V.
A crystal current of 50uA to 1mA or so might perhaps be expected for a
typical 10MHz overtone crystal.
Unless some non standard crystal cut with exceptionally high ESR is used.
The oscillator transistor dc emitter current appears consistent with a
crystal current of around 1mA or so.
Can use Spice to check but there is no need for elaborate starting
techniques when simulating this circuit as it has plenty of excess gain
if the simulated Q isn't too large (50,000 or so is easy) (only ESR has
much effect on the crystal current).
Elaborate simulations looking for ppm changes when the active device
models aren't that good can be counterproductive.
Doing this with any confidence requires enormous trust in the
convergence of the numerical algorithms employed.
All flavours of Spice do have problems with accuracy and convergence
when the condition number of the solution matrix is large enough.
Often this is because the simulated circuit is unstable so an educated
guess as to what to change to preserve stability/convergence is required.
Occasionally Spice can even predict noise that is much larger than that
actually measured (it can have trouble when one has both positive and
negative feedback loops).
Bruce
WarrenS wrote:
>> Thus one cannot arbitrarily increase Vcc without considering the consequences ...
>>
>
> True, but with the AC voltages as low as they are, I do think it is safe to think that the crystal current should be higher.
> I was not suggesting to arbitrarily rise the Vcc, I was asking any chance it is at the wrong voltage???
> Even though 2.49 volts is plenty of Vcc to design an osc circuit to run on, It seems like a low, funny voltage to design for, on what I assume is an old product.
>
> ws
> *******************
> ----- Original Message -----
> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
> Sent: Friday, July 03, 2009 8:36 PM
> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>
>
>
>> WB6BNQ wrote:
>>
>>> I agree with Warren. The VCC seems way too low. Interestingly, noone else seems to think so,
>>> or did not notice it.
>>>
>>> Bill....WB6BNQ
>>>
>>>
>>>
>>>
>> It was noticed, but it is consistent with the resistor values used in
>> the circuit.
>> The crystal current is determined by the dc emitter current of the
>> oscillator transistor.
>> Thus one cannot arbitrarily increase Vcc without considering the
>> consequences for the crystal current etc.
>>
>> Bruce
>>
>>> WarrenS wrote:
>>>
>>>
>>>
>>>> Here is my two cents worth
>>>> 20 mv output, sure sounds like something is broken.
>>>> It should be fixed before it is modified.
>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>> If you do 'need' to modify the gain,
>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>
>>>> ws
>>>>
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>>
>>
>>
>>
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
W
WarrenS
Sat, Jul 4, 2009 5:54 AM
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
A couple more thoughts;
Is the VCC (Red wire) current draw correct?.
Should be 1.3 ma (per the schematic values) with the 470 Ohm and
more like 7 plus ma with the 47 ohm.
AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
If both answers are YES, at least most things are working correctly.
As Bruce said:
"if the resistor values are correct this is the voltage(s) that one should get"
With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
AND the output amplitude is not yet being controlled by current cut off.
Something that needs to be asked is are all these the correct values.
It sounds like you got them by measuring and not from a "should be" schematic.
One last, way out thought.
It was my belief that very good osc have some sort of AGC to control the amplitude,
ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
ws
The Colpitts crystal oscillator limits the crystal current by being
periodically cutoff for part of the cycle.
There is an optimum duty cycle (usually something like 25%) that
corresponds to minimum phase noise.
The oscillator dc collector current determines the crystal current and
hence the output amplitude.
It is possible to perform AGC by varying the oscillator Vcc but the
output amplifier schematic appears to have no circuitry for this.
Some circuits actually do this, but the oscillator transistor is still
cutoff for part of the oscillator cycle.
The oscillator VCC is fed from a tap in a resistive divider network
(located on the external (to the oven) buffer board) connected between
the 5V supply and ground.
Usually one just varies the emitter (or collector) dc current by
selecting a resistor value.
Bruce
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:21 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
Not so, if the resistor values are correct this is the voltage that one
should get.
The question that remains: Is the voltage what the designer intended?
With the Colpitts oscillator used limiting in the oscillator transistor
(usually by current cutoff) is used to limit the amplitude.
Just increasing the voltage without checking the resultant crystal
current will be within permissable limits can be counter productive.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
An update on the Broken Ovenaire OSC 85-50.
I prepared a 'schematic' of the Output Board and the Oscillator
Board (attached) and I have lots of pictures of the external unit
and the insides if anyone is interested.
I resoldered all connections and replaced all transistors on the
Output Board and the Oscillator Board all to no benefit. I
measured all the components with an LCR meter and found the 0.01
uF bypass on the 330 ohm resistor in the emitter circuit of the
output transistor of the Output Board to be low and with a high
ESR. I replaced this with about a 20% increase in output amplitude
but still inadequate. I replaced the rest of the 0.01 uF caps on
the output board with no additional benefit. I transiently
disconnected the Red wires from the Oven Controller board and
there was no increase in output or significant increase in voltage
to the Oscillator Board.
Therefore, it appeared that a 'low output crystal' (if such a
thing exists) was the only logical explanation that I could come
up with. That seeming to be the case, there appeared to be only 4
options. 1. Toss the OCXO (sorry, too much effort so far). 2.
Build an external amplifier (seemingly too much additional
effort). 3. Try to adjust on the bias of the oscillator transistor
to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
value of the resistor in the emitter circuit of the Oscillator
Board to get more gain out of the last stage in the Oscillator
Board.
I replaced the 470 ohm resistor with a 47 ohm resistor and the
amplitude increased to about 0.4 V P-P into a 50 ohm load and was
sufficient to make it a usable OCXO again.
I reassembled, resealed with Epoxy and all seems well so far.
If anyone wants pictures or other info, please let me know.
Thanks for all the suggestions and help.
Joe,
Congratulations on getting your system to work!
A couple of things. First, trying to measure the currents in the
circuit with a ferrite toroid won't do you much good. You don't know
what the currents should be, and the secondary of the toroid
transformer requires a termination resistor. The value changes with
the turns ratio.
Just from looking at the circuit, the RF currents will be extremely
low. This requires a large number of turns on the secondary, which
will probably resonate at or below the 10MHz operating frequency due
to stray capacitance from the connection to the scope. So it is
unlikely you will get any useful progress in this direction.
However, from the values on your schematic, the output tank circuit
resonates at 9.602MHz with a Q of 9.6. So the tank is already well
below resonance, which attenuates the output voltage.
Any stray capacitance you add to the circuit will bring the resonant
frequency lower, further aggravating the loss in signal.
The output tank is tapped with the 75pF and 91pF in series. This
further attenuates the signal.
I'd change the circuit to a single capacitor across the tank with a
small trim capacitor to tune it to resonance.
To get the signal into 50 ohms for distribution, I'd add a limiter
if you can tolerate a square wave output, or a good emitter follower
if you need a sine wave. Take the output from the collector of the
2N2369 to get the maximum signal amplitude.
Your original post mentions an output amplitude of 20mV. If the
normal amplitude is around 2V, this represents a loss of 40dB. This
is a huge loss in signal. The circuit obviously worked at one time,
so there may well be some other hidden problem.
It is possible the crystal is damaged, but this seems unlikely. A
crystal oscillator probably won't even start if the signal level is
down 40dB.
You can check the oscillator and crystal in SPICE. Normally, the
high Q of the crystal will make the analysis very slow. It could
take many hours for the simulation to begin oscillating and
stabilize at the final amplitude. The transient analysis requires a
very fine time step for accuracy, and you could run out of memory
before the simulation was complete.
I have developed a much faster way of analyzing a crystal oscillator
in SPICE. Instead of requiring tens or hundreds of thousands of
simulated cycles, this method gives accurate results in only a few
dozen cycles. For more information, please see "SPICE Analysis of
Crystal Oscillators"
http://pstca.com/spice/xtal/clapp.htm
You can estimate the value of the crystal ESR by finding the Q of
your crystal and working backwards.
I'm attaching a gif of your schematic for reference. This is rotated
90 degrees and enhanced in LView Pro to improve the contrast.
Please let me know if you have any questions.
Thanks,
Mike
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
****************
----- Original Message -----
From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
> WarrenS wrote:
>> A couple more thoughts;
>>
>> Is the VCC (Red wire) current draw correct?.
>> Should be 1.3 ma (per the schematic values) with the 470 Ohm and
>> more like 7 plus ma with the 47 ohm.
>>
>> AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
>> If both answers are YES, at least most things are working correctly.
>>
>> As Bruce said:
>>
>>> "if the resistor values are correct this is the voltage(s) that one should get"
>>>
>> With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
>> BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
>> AND the output amplitude is not yet being controlled by current cut off.
>>
>> Something that needs to be asked is are all these the correct values.
>> It sounds like you got them by measuring and not from a "should be" schematic.
>>
>> One last, way out thought.
>> It was my belief that very good osc have some sort of AGC to control the amplitude,
>> ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
>> If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
>>
>> ws
>>
>>
> The Colpitts crystal oscillator limits the crystal current by being
> periodically cutoff for part of the cycle.
> There is an optimum duty cycle (usually something like 25%) that
> corresponds to minimum phase noise.
> The oscillator dc collector current determines the crystal current and
> hence the output amplitude.
> It is possible to perform AGC by varying the oscillator Vcc but the
> output amplifier schematic appears to have no circuitry for this.
> Some circuits actually do this, but the oscillator transistor is still
> cutoff for part of the oscillator cycle.
> The oscillator VCC is fed from a tap in a resistive divider network
> (located on the external (to the oven) buffer board) connected between
> the 5V supply and ground.
> Usually one just varies the emitter (or collector) dc current by
> selecting a resistor value.
>
> Bruce
>> *****************
>> ----- Original Message -----
>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>> Sent: Friday, July 03, 2009 8:21 PM
>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>
>>
>>
>>> WarrenS wrote:
>>>
>>>> Here is my two cents worth
>>>> 20 mv output, sure sounds like something is broken.
>>>> It should be fixed before it is modified.
>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>>
>>>>
>>> Not so, if the resistor values are correct this is the voltage that one
>>> should get.
>>> The question that remains: Is the voltage what the designer intended?
>>> With the Colpitts oscillator used limiting in the oscillator transistor
>>> (usually by current cutoff) is used to limit the amplitude.
>>> Just increasing the voltage without checking the resultant crystal
>>> current will be within permissable limits can be counter productive.
>>>
>>>
>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>> If you do 'need' to modify the gain,
>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>
>>>> ws
>>>>
>>>>
>>> Bruce
>>>
>>>> ********************
>>>> ----- Original Message -----
>>>> From: "Mike Monett" <xde-l2g3@myamail.com>
>>>> To: <time-nuts@febo.com>
>>>> Sent: Friday, July 03, 2009 11:50 AM
>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>
>>>>
>>>>
>>>>
>>>>> > An update on the Broken Ovenaire OSC 85-50.
>>>>>
>>>>> > I prepared a 'schematic' of the Output Board and the Oscillator
>>>>> > Board (attached) and I have lots of pictures of the external unit
>>>>> > and the insides if anyone is interested.
>>>>>
>>>>> > I resoldered all connections and replaced all transistors on the
>>>>> > Output Board and the Oscillator Board all to no benefit. I
>>>>> > measured all the components with an LCR meter and found the 0.01
>>>>> > uF bypass on the 330 ohm resistor in the emitter circuit of the
>>>>> > output transistor of the Output Board to be low and with a high
>>>>> > ESR. I replaced this with about a 20% increase in output amplitude
>>>>> > but still inadequate. I replaced the rest of the 0.01 uF caps on
>>>>> > the output board with no additional benefit. I transiently
>>>>> > disconnected the Red wires from the Oven Controller board and
>>>>> > there was no increase in output or significant increase in voltage
>>>>> > to the Oscillator Board.
>>>>>
>>>>> > Therefore, it appeared that a 'low output crystal' (if such a
>>>>> > thing exists) was the only logical explanation that I could come
>>>>> > up with. That seeming to be the case, there appeared to be only 4
>>>>> > options. 1. Toss the OCXO (sorry, too much effort so far). 2.
>>>>>
>>>>> > Build an external amplifier (seemingly too much additional
>>>>> > effort). 3. Try to adjust on the bias of the oscillator transistor
>>>>> > to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
>>>>> > value of the resistor in the emitter circuit of the Oscillator
>>>>> > Board to get more gain out of the last stage in the Oscillator
>>>>> > Board.
>>>>>
>>>>> > I replaced the 470 ohm resistor with a 47 ohm resistor and the
>>>>> > amplitude increased to about 0.4 V P-P into a 50 ohm load and was
>>>>> > sufficient to make it a usable OCXO again.
>>>>>
>>>>> > I reassembled, resealed with Epoxy and all seems well so far.
>>>>>
>>>>> > If anyone wants pictures or other info, please let me know.
>>>>>
>>>>> > Thanks for all the suggestions and help.
>>>>>
>>>>> > Joe
>>>>>
>>>>> Joe,
>>>>>
>>>>> Congratulations on getting your system to work!
>>>>>
>>>>> A couple of things. First, trying to measure the currents in the
>>>>> circuit with a ferrite toroid won't do you much good. You don't know
>>>>> what the currents should be, and the secondary of the toroid
>>>>> transformer requires a termination resistor. The value changes with
>>>>> the turns ratio.
>>>>>
>>>>> Just from looking at the circuit, the RF currents will be extremely
>>>>> low. This requires a large number of turns on the secondary, which
>>>>> will probably resonate at or below the 10MHz operating frequency due
>>>>> to stray capacitance from the connection to the scope. So it is
>>>>> unlikely you will get any useful progress in this direction.
>>>>>
>>>>> However, from the values on your schematic, the output tank circuit
>>>>> resonates at 9.602MHz with a Q of 9.6. So the tank is already well
>>>>> below resonance, which attenuates the output voltage.
>>>>>
>>>>> Any stray capacitance you add to the circuit will bring the resonant
>>>>> frequency lower, further aggravating the loss in signal.
>>>>>
>>>>> The output tank is tapped with the 75pF and 91pF in series. This
>>>>> further attenuates the signal.
>>>>>
>>>>> I'd change the circuit to a single capacitor across the tank with a
>>>>> small trim capacitor to tune it to resonance.
>>>>>
>>>>> To get the signal into 50 ohms for distribution, I'd add a limiter
>>>>> if you can tolerate a square wave output, or a good emitter follower
>>>>> if you need a sine wave. Take the output from the collector of the
>>>>> 2N2369 to get the maximum signal amplitude.
>>>>>
>>>>> Your original post mentions an output amplitude of 20mV. If the
>>>>> normal amplitude is around 2V, this represents a loss of 40dB. This
>>>>> is a huge loss in signal. The circuit obviously worked at one time,
>>>>> so there may well be some other hidden problem.
>>>>>
>>>>> It is possible the crystal is damaged, but this seems unlikely. A
>>>>> crystal oscillator probably won't even start if the signal level is
>>>>> down 40dB.
>>>>>
>>>>> You can check the oscillator and crystal in SPICE. Normally, the
>>>>> high Q of the crystal will make the analysis very slow. It could
>>>>> take many hours for the simulation to begin oscillating and
>>>>> stabilize at the final amplitude. The transient analysis requires a
>>>>> very fine time step for accuracy, and you could run out of memory
>>>>> before the simulation was complete.
>>>>>
>>>>> I have developed a much faster way of analyzing a crystal oscillator
>>>>> in SPICE. Instead of requiring tens or hundreds of thousands of
>>>>> simulated cycles, this method gives accurate results in only a few
>>>>> dozen cycles. For more information, please see "SPICE Analysis of
>>>>> Crystal Oscillators"
>>>>>
>>>>> http://pstca.com/spice/xtal/clapp.htm
>>>>>
>>>>> You can estimate the value of the crystal ESR by finding the Q of
>>>>> your crystal and working backwards.
>>>>>
>>>>> I'm attaching a gif of your schematic for reference. This is rotated
>>>>> 90 degrees and enhanced in LView Pro to improve the contrast.
>>>>>
>>>>> Please let me know if you have any questions.
>>>>>
>>>>> Thanks,
>>>>>
>>>>> Mike
>>>>>
>>>>>
>>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>>
>>>>
>>>
>>>
>>>
>>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
>
>
>
>
BG
Bruce Griffiths
Sat, Jul 4, 2009 6:08 AM
Warren
My Spice model of the dc levels in the oscillator indicate a current
more like 4mA in the 510 ohm resistor.
Bruce
WarrenS wrote:
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
A couple more thoughts;
Is the VCC (Red wire) current draw correct?.
Should be 1.3 ma (per the schematic values) with the 470 Ohm and
more like 7 plus ma with the 47 ohm.
AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
If both answers are YES, at least most things are working correctly.
As Bruce said:
"if the resistor values are correct this is the voltage(s) that one should get"
With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
AND the output amplitude is not yet being controlled by current cut off.
Something that needs to be asked is are all these the correct values.
It sounds like you got them by measuring and not from a "should be" schematic.
One last, way out thought.
It was my belief that very good osc have some sort of AGC to control the amplitude,
ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
ws
The Colpitts crystal oscillator limits the crystal current by being
periodically cutoff for part of the cycle.
There is an optimum duty cycle (usually something like 25%) that
corresponds to minimum phase noise.
The oscillator dc collector current determines the crystal current and
hence the output amplitude.
It is possible to perform AGC by varying the oscillator Vcc but the
output amplifier schematic appears to have no circuitry for this.
Some circuits actually do this, but the oscillator transistor is still
cutoff for part of the oscillator cycle.
The oscillator VCC is fed from a tap in a resistive divider network
(located on the external (to the oven) buffer board) connected between
the 5V supply and ground.
Usually one just varies the emitter (or collector) dc current by
selecting a resistor value.
Bruce
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:21 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
Not so, if the resistor values are correct this is the voltage that one
should get.
The question that remains: Is the voltage what the designer intended?
With the Colpitts oscillator used limiting in the oscillator transistor
(usually by current cutoff) is used to limit the amplitude.
Just increasing the voltage without checking the resultant crystal
current will be within permissable limits can be counter productive.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
An update on the Broken Ovenaire OSC 85-50.
I prepared a 'schematic' of the Output Board and the Oscillator
Board (attached) and I have lots of pictures of the external unit
and the insides if anyone is interested.
I resoldered all connections and replaced all transistors on the
Output Board and the Oscillator Board all to no benefit. I
measured all the components with an LCR meter and found the 0.01
uF bypass on the 330 ohm resistor in the emitter circuit of the
output transistor of the Output Board to be low and with a high
ESR. I replaced this with about a 20% increase in output amplitude
but still inadequate. I replaced the rest of the 0.01 uF caps on
the output board with no additional benefit. I transiently
disconnected the Red wires from the Oven Controller board and
there was no increase in output or significant increase in voltage
to the Oscillator Board.
Therefore, it appeared that a 'low output crystal' (if such a
thing exists) was the only logical explanation that I could come
up with. That seeming to be the case, there appeared to be only 4
options. 1. Toss the OCXO (sorry, too much effort so far). 2.
Build an external amplifier (seemingly too much additional
effort). 3. Try to adjust on the bias of the oscillator transistor
to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
value of the resistor in the emitter circuit of the Oscillator
Board to get more gain out of the last stage in the Oscillator
Board.
I replaced the 470 ohm resistor with a 47 ohm resistor and the
amplitude increased to about 0.4 V P-P into a 50 ohm load and was
sufficient to make it a usable OCXO again.
I reassembled, resealed with Epoxy and all seems well so far.
If anyone wants pictures or other info, please let me know.
Thanks for all the suggestions and help.
Joe,
Congratulations on getting your system to work!
A couple of things. First, trying to measure the currents in the
circuit with a ferrite toroid won't do you much good. You don't know
what the currents should be, and the secondary of the toroid
transformer requires a termination resistor. The value changes with
the turns ratio.
Just from looking at the circuit, the RF currents will be extremely
low. This requires a large number of turns on the secondary, which
will probably resonate at or below the 10MHz operating frequency due
to stray capacitance from the connection to the scope. So it is
unlikely you will get any useful progress in this direction.
However, from the values on your schematic, the output tank circuit
resonates at 9.602MHz with a Q of 9.6. So the tank is already well
below resonance, which attenuates the output voltage.
Any stray capacitance you add to the circuit will bring the resonant
frequency lower, further aggravating the loss in signal.
The output tank is tapped with the 75pF and 91pF in series. This
further attenuates the signal.
I'd change the circuit to a single capacitor across the tank with a
small trim capacitor to tune it to resonance.
To get the signal into 50 ohms for distribution, I'd add a limiter
if you can tolerate a square wave output, or a good emitter follower
if you need a sine wave. Take the output from the collector of the
2N2369 to get the maximum signal amplitude.
Your original post mentions an output amplitude of 20mV. If the
normal amplitude is around 2V, this represents a loss of 40dB. This
is a huge loss in signal. The circuit obviously worked at one time,
so there may well be some other hidden problem.
It is possible the crystal is damaged, but this seems unlikely. A
crystal oscillator probably won't even start if the signal level is
down 40dB.
You can check the oscillator and crystal in SPICE. Normally, the
high Q of the crystal will make the analysis very slow. It could
take many hours for the simulation to begin oscillating and
stabilize at the final amplitude. The transient analysis requires a
very fine time step for accuracy, and you could run out of memory
before the simulation was complete.
I have developed a much faster way of analyzing a crystal oscillator
in SPICE. Instead of requiring tens or hundreds of thousands of
simulated cycles, this method gives accurate results in only a few
dozen cycles. For more information, please see "SPICE Analysis of
Crystal Oscillators"
http://pstca.com/spice/xtal/clapp.htm
You can estimate the value of the crystal ESR by finding the Q of
your crystal and working backwards.
I'm attaching a gif of your schematic for reference. This is rotated
90 degrees and enhanced in LView Pro to improve the contrast.
Please let me know if you have any questions.
Thanks,
Mike
Warren
My Spice model of the dc levels in the oscillator indicate a current
more like 4mA in the 510 ohm resistor.
Bruce
WarrenS wrote:
> I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
>
> The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
> Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
> If I'm correct about that (this time) then, It has got to be one of three things.
> The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
> It should be easy enough to measure which one it is.
> The voltage at the junction of the two 4.7K resistors are WRONG.
> Looks like it is caused by an excess current draw on the Grn lead.
> Check the direction and leakage of the voltage controlled cap diode and bypass.
> If all else is right, Look for the cause of the extra loading current on the 510 ohm,
>
> WS
>
> ****************
>
> ----- Original Message -----
> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
> Sent: Friday, July 03, 2009 9:55 PM
> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>
>
>
>> WarrenS wrote:
>>
>>> A couple more thoughts;
>>>
>>> Is the VCC (Red wire) current draw correct?.
>>> Should be 1.3 ma (per the schematic values) with the 470 Ohm and
>>> more like 7 plus ma with the 47 ohm.
>>>
>>> AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
>>> If both answers are YES, at least most things are working correctly.
>>>
>>> As Bruce said:
>>>
>>>
>>>> "if the resistor values are correct this is the voltage(s) that one should get"
>>>>
>>>>
>>> With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
>>> BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
>>> AND the output amplitude is not yet being controlled by current cut off.
>>>
>>> Something that needs to be asked is are all these the correct values.
>>> It sounds like you got them by measuring and not from a "should be" schematic.
>>>
>>> One last, way out thought.
>>> It was my belief that very good osc have some sort of AGC to control the amplitude,
>>> ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
>>> If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
>>>
>>> ws
>>>
>>>
>>>
>> The Colpitts crystal oscillator limits the crystal current by being
>> periodically cutoff for part of the cycle.
>> There is an optimum duty cycle (usually something like 25%) that
>> corresponds to minimum phase noise.
>> The oscillator dc collector current determines the crystal current and
>> hence the output amplitude.
>> It is possible to perform AGC by varying the oscillator Vcc but the
>> output amplifier schematic appears to have no circuitry for this.
>> Some circuits actually do this, but the oscillator transistor is still
>> cutoff for part of the oscillator cycle.
>> The oscillator VCC is fed from a tap in a resistive divider network
>> (located on the external (to the oven) buffer board) connected between
>> the 5V supply and ground.
>> Usually one just varies the emitter (or collector) dc current by
>> selecting a resistor value.
>>
>> Bruce
>>
>>> *****************
>>> ----- Original Message -----
>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>> Sent: Friday, July 03, 2009 8:21 PM
>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>
>>>
>>>
>>>
>>>> WarrenS wrote:
>>>>
>>>>
>>>>> Here is my two cents worth
>>>>> 20 mv output, sure sounds like something is broken.
>>>>> It should be fixed before it is modified.
>>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>>>
>>>>>
>>>>>
>>>> Not so, if the resistor values are correct this is the voltage that one
>>>> should get.
>>>> The question that remains: Is the voltage what the designer intended?
>>>> With the Colpitts oscillator used limiting in the oscillator transistor
>>>> (usually by current cutoff) is used to limit the amplitude.
>>>> Just increasing the voltage without checking the resultant crystal
>>>> current will be within permissable limits can be counter productive.
>>>>
>>>>
>>>>
>>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>>> If you do 'need' to modify the gain,
>>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>>
>>>>> ws
>>>>>
>>>>>
>>>>>
>>>> Bruce
>>>>
>>>>
>>>>> ********************
>>>>> ----- Original Message -----
>>>>> From: "Mike Monett" <xde-l2g3@myamail.com>
>>>>> To: <time-nuts@febo.com>
>>>>> Sent: Friday, July 03, 2009 11:50 AM
>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>> > An update on the Broken Ovenaire OSC 85-50.
>>>>>>
>>>>>> > I prepared a 'schematic' of the Output Board and the Oscillator
>>>>>> > Board (attached) and I have lots of pictures of the external unit
>>>>>> > and the insides if anyone is interested.
>>>>>>
>>>>>> > I resoldered all connections and replaced all transistors on the
>>>>>> > Output Board and the Oscillator Board all to no benefit. I
>>>>>> > measured all the components with an LCR meter and found the 0.01
>>>>>> > uF bypass on the 330 ohm resistor in the emitter circuit of the
>>>>>> > output transistor of the Output Board to be low and with a high
>>>>>> > ESR. I replaced this with about a 20% increase in output amplitude
>>>>>> > but still inadequate. I replaced the rest of the 0.01 uF caps on
>>>>>> > the output board with no additional benefit. I transiently
>>>>>> > disconnected the Red wires from the Oven Controller board and
>>>>>> > there was no increase in output or significant increase in voltage
>>>>>> > to the Oscillator Board.
>>>>>>
>>>>>> > Therefore, it appeared that a 'low output crystal' (if such a
>>>>>> > thing exists) was the only logical explanation that I could come
>>>>>> > up with. That seeming to be the case, there appeared to be only 4
>>>>>> > options. 1. Toss the OCXO (sorry, too much effort so far). 2.
>>>>>>
>>>>>> > Build an external amplifier (seemingly too much additional
>>>>>> > effort). 3. Try to adjust on the bias of the oscillator transistor
>>>>>> > to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
>>>>>> > value of the resistor in the emitter circuit of the Oscillator
>>>>>> > Board to get more gain out of the last stage in the Oscillator
>>>>>> > Board.
>>>>>>
>>>>>> > I replaced the 470 ohm resistor with a 47 ohm resistor and the
>>>>>> > amplitude increased to about 0.4 V P-P into a 50 ohm load and was
>>>>>> > sufficient to make it a usable OCXO again.
>>>>>>
>>>>>> > I reassembled, resealed with Epoxy and all seems well so far.
>>>>>>
>>>>>> > If anyone wants pictures or other info, please let me know.
>>>>>>
>>>>>> > Thanks for all the suggestions and help.
>>>>>>
>>>>>> > Joe
>>>>>>
>>>>>> Joe,
>>>>>>
>>>>>> Congratulations on getting your system to work!
>>>>>>
>>>>>> A couple of things. First, trying to measure the currents in the
>>>>>> circuit with a ferrite toroid won't do you much good. You don't know
>>>>>> what the currents should be, and the secondary of the toroid
>>>>>> transformer requires a termination resistor. The value changes with
>>>>>> the turns ratio.
>>>>>>
>>>>>> Just from looking at the circuit, the RF currents will be extremely
>>>>>> low. This requires a large number of turns on the secondary, which
>>>>>> will probably resonate at or below the 10MHz operating frequency due
>>>>>> to stray capacitance from the connection to the scope. So it is
>>>>>> unlikely you will get any useful progress in this direction.
>>>>>>
>>>>>> However, from the values on your schematic, the output tank circuit
>>>>>> resonates at 9.602MHz with a Q of 9.6. So the tank is already well
>>>>>> below resonance, which attenuates the output voltage.
>>>>>>
>>>>>> Any stray capacitance you add to the circuit will bring the resonant
>>>>>> frequency lower, further aggravating the loss in signal.
>>>>>>
>>>>>> The output tank is tapped with the 75pF and 91pF in series. This
>>>>>> further attenuates the signal.
>>>>>>
>>>>>> I'd change the circuit to a single capacitor across the tank with a
>>>>>> small trim capacitor to tune it to resonance.
>>>>>>
>>>>>> To get the signal into 50 ohms for distribution, I'd add a limiter
>>>>>> if you can tolerate a square wave output, or a good emitter follower
>>>>>> if you need a sine wave. Take the output from the collector of the
>>>>>> 2N2369 to get the maximum signal amplitude.
>>>>>>
>>>>>> Your original post mentions an output amplitude of 20mV. If the
>>>>>> normal amplitude is around 2V, this represents a loss of 40dB. This
>>>>>> is a huge loss in signal. The circuit obviously worked at one time,
>>>>>> so there may well be some other hidden problem.
>>>>>>
>>>>>> It is possible the crystal is damaged, but this seems unlikely. A
>>>>>> crystal oscillator probably won't even start if the signal level is
>>>>>> down 40dB.
>>>>>>
>>>>>> You can check the oscillator and crystal in SPICE. Normally, the
>>>>>> high Q of the crystal will make the analysis very slow. It could
>>>>>> take many hours for the simulation to begin oscillating and
>>>>>> stabilize at the final amplitude. The transient analysis requires a
>>>>>> very fine time step for accuracy, and you could run out of memory
>>>>>> before the simulation was complete.
>>>>>>
>>>>>> I have developed a much faster way of analyzing a crystal oscillator
>>>>>> in SPICE. Instead of requiring tens or hundreds of thousands of
>>>>>> simulated cycles, this method gives accurate results in only a few
>>>>>> dozen cycles. For more information, please see "SPICE Analysis of
>>>>>> Crystal Oscillators"
>>>>>>
>>>>>> http://pstca.com/spice/xtal/clapp.htm
>>>>>>
>>>>>> You can estimate the value of the crystal ESR by finding the Q of
>>>>>> your crystal and working backwards.
>>>>>>
>>>>>> I'm attaching a gif of your schematic for reference. This is rotated
>>>>>> 90 degrees and enhanced in LView Pro to improve the contrast.
>>>>>>
>>>>>> Please let me know if you have any questions.
>>>>>>
>>>>>> Thanks,
>>>>>>
>>>>>> Mike
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>>
>>>>>
>>>>>
>>>>
>>>>
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>>
>>
>>
>>
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
W
WarrenS
Sat, Jul 4, 2009 6:37 AM
Bruce
4ma? Interesting, Maybe IF the Osc was running, BUT
My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
With it NOT osc, which is almost the present case,
the Only Dc current drawn is.
1.75 V on 20K = 0.08ma (Q1 bias)
1.04 V on 2.2K = 0.47ma (Q1 collector)
0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
Osc board =1.32 ma
Buffer board
(2*4.7k) at 2.49 V = 0.26 ma
Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 11:08 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Warren
My Spice model of the dc levels in the oscillator indicate a current
more like 4mA in the 510 ohm resistor.
Bruce
WarrenS wrote:
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
A couple more thoughts;
Is the VCC (Red wire) current draw correct?.
Should be 1.3 ma (per the schematic values) with the 470 Ohm and
more like 7 plus ma with the 47 ohm.
AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
If both answers are YES, at least most things are working correctly.
As Bruce said:
"if the resistor values are correct this is the voltage(s) that one should get"
With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
AND the output amplitude is not yet being controlled by current cut off.
Something that needs to be asked is are all these the correct values.
It sounds like you got them by measuring and not from a "should be" schematic.
One last, way out thought.
It was my belief that very good osc have some sort of AGC to control the amplitude,
ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
ws
The Colpitts crystal oscillator limits the crystal current by being
periodically cutoff for part of the cycle.
There is an optimum duty cycle (usually something like 25%) that
corresponds to minimum phase noise.
The oscillator dc collector current determines the crystal current and
hence the output amplitude.
It is possible to perform AGC by varying the oscillator Vcc but the
output amplifier schematic appears to have no circuitry for this.
Some circuits actually do this, but the oscillator transistor is still
cutoff for part of the oscillator cycle.
The oscillator VCC is fed from a tap in a resistive divider network
(located on the external (to the oven) buffer board) connected between
the 5V supply and ground.
Usually one just varies the emitter (or collector) dc current by
selecting a resistor value.
Bruce
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:21 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
Not so, if the resistor values are correct this is the voltage that one
should get.
The question that remains: Is the voltage what the designer intended?
With the Colpitts oscillator used limiting in the oscillator transistor
(usually by current cutoff) is used to limit the amplitude.
Just increasing the voltage without checking the resultant crystal
current will be within permissable limits can be counter productive.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
An update on the Broken Ovenaire OSC 85-50.
I prepared a 'schematic' of the Output Board and the Oscillator
Board (attached) and I have lots of pictures of the external unit
and the insides if anyone is interested.
I resoldered all connections and replaced all transistors on the
Output Board and the Oscillator Board all to no benefit. I
measured all the components with an LCR meter and found the 0.01
uF bypass on the 330 ohm resistor in the emitter circuit of the
output transistor of the Output Board to be low and with a high
ESR. I replaced this with about a 20% increase in output amplitude
but still inadequate. I replaced the rest of the 0.01 uF caps on
the output board with no additional benefit. I transiently
disconnected the Red wires from the Oven Controller board and
there was no increase in output or significant increase in voltage
to the Oscillator Board.
Therefore, it appeared that a 'low output crystal' (if such a
thing exists) was the only logical explanation that I could come
up with. That seeming to be the case, there appeared to be only 4
options. 1. Toss the OCXO (sorry, too much effort so far). 2.
Build an external amplifier (seemingly too much additional
effort). 3. Try to adjust on the bias of the oscillator transistor
to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
value of the resistor in the emitter circuit of the Oscillator
Board to get more gain out of the last stage in the Oscillator
Board.
I replaced the 470 ohm resistor with a 47 ohm resistor and the
amplitude increased to about 0.4 V P-P into a 50 ohm load and was
sufficient to make it a usable OCXO again.
I reassembled, resealed with Epoxy and all seems well so far.
If anyone wants pictures or other info, please let me know.
Thanks for all the suggestions and help.
Joe,
Congratulations on getting your system to work!
A couple of things. First, trying to measure the currents in the
circuit with a ferrite toroid won't do you much good. You don't know
what the currents should be, and the secondary of the toroid
transformer requires a termination resistor. The value changes with
the turns ratio.
Just from looking at the circuit, the RF currents will be extremely
low. This requires a large number of turns on the secondary, which
will probably resonate at or below the 10MHz operating frequency due
to stray capacitance from the connection to the scope. So it is
unlikely you will get any useful progress in this direction.
However, from the values on your schematic, the output tank circuit
resonates at 9.602MHz with a Q of 9.6. So the tank is already well
below resonance, which attenuates the output voltage.
Any stray capacitance you add to the circuit will bring the resonant
frequency lower, further aggravating the loss in signal.
The output tank is tapped with the 75pF and 91pF in series. This
further attenuates the signal.
I'd change the circuit to a single capacitor across the tank with a
small trim capacitor to tune it to resonance.
To get the signal into 50 ohms for distribution, I'd add a limiter
if you can tolerate a square wave output, or a good emitter follower
if you need a sine wave. Take the output from the collector of the
2N2369 to get the maximum signal amplitude.
Your original post mentions an output amplitude of 20mV. If the
normal amplitude is around 2V, this represents a loss of 40dB. This
is a huge loss in signal. The circuit obviously worked at one time,
so there may well be some other hidden problem.
It is possible the crystal is damaged, but this seems unlikely. A
crystal oscillator probably won't even start if the signal level is
down 40dB.
You can check the oscillator and crystal in SPICE. Normally, the
high Q of the crystal will make the analysis very slow. It could
take many hours for the simulation to begin oscillating and
stabilize at the final amplitude. The transient analysis requires a
very fine time step for accuracy, and you could run out of memory
before the simulation was complete.
I have developed a much faster way of analyzing a crystal oscillator
in SPICE. Instead of requiring tens or hundreds of thousands of
simulated cycles, this method gives accurate results in only a few
dozen cycles. For more information, please see "SPICE Analysis of
Crystal Oscillators"
http://pstca.com/spice/xtal/clapp.htm
You can estimate the value of the crystal ESR by finding the Q of
your crystal and working backwards.
I'm attaching a gif of your schematic for reference. This is rotated
90 degrees and enhanced in LView Pro to improve the contrast.
Please let me know if you have any questions.
Thanks,
Mike
Bruce
4ma? Interesting, Maybe IF the Osc was running, BUT
My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
With it NOT osc, which is almost the present case,
the Only Dc current drawn is.
1.75 V on 20K = 0.08ma (Q1 bias)
1.04 V on 2.2K = 0.47ma (Q1 collector)
0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
Osc board =1.32 ma
Buffer board
(2*4.7k) at 2.49 V = 0.26 ma
Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
ws
**************
----- Original Message -----
From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
Sent: Friday, July 03, 2009 11:08 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
> Warren
>
> My Spice model of the dc levels in the oscillator indicate a current
> more like 4mA in the 510 ohm resistor.
>
> Bruce
>
> WarrenS wrote:
>> I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
>>
>> The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
>> Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
>> If I'm correct about that (this time) then, It has got to be one of three things.
>> The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
>> It should be easy enough to measure which one it is.
>> The voltage at the junction of the two 4.7K resistors are WRONG.
>> Looks like it is caused by an excess current draw on the Grn lead.
>> Check the direction and leakage of the voltage controlled cap diode and bypass.
>> If all else is right, Look for the cause of the extra loading current on the 510 ohm,
>>
>> WS
>>
>> ****************
>>
>> ----- Original Message -----
>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>> Sent: Friday, July 03, 2009 9:55 PM
>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>
>>
>>
>>> WarrenS wrote:
>>>
>>>> A couple more thoughts;
>>>>
>>>> Is the VCC (Red wire) current draw correct?.
>>>> Should be 1.3 ma (per the schematic values) with the 470 Ohm and
>>>> more like 7 plus ma with the 47 ohm.
>>>>
>>>> AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
>>>> If both answers are YES, at least most things are working correctly.
>>>>
>>>> As Bruce said:
>>>>
>>>>
>>>>> "if the resistor values are correct this is the voltage(s) that one should get"
>>>>>
>>>>>
>>>> With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
>>>> BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
>>>> AND the output amplitude is not yet being controlled by current cut off.
>>>>
>>>> Something that needs to be asked is are all these the correct values.
>>>> It sounds like you got them by measuring and not from a "should be" schematic.
>>>>
>>>> One last, way out thought.
>>>> It was my belief that very good osc have some sort of AGC to control the amplitude,
>>>> ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
>>>> If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
>>>>
>>>> ws
>>>>
>>>>
>>>>
>>> The Colpitts crystal oscillator limits the crystal current by being
>>> periodically cutoff for part of the cycle.
>>> There is an optimum duty cycle (usually something like 25%) that
>>> corresponds to minimum phase noise.
>>> The oscillator dc collector current determines the crystal current and
>>> hence the output amplitude.
>>> It is possible to perform AGC by varying the oscillator Vcc but the
>>> output amplifier schematic appears to have no circuitry for this.
>>> Some circuits actually do this, but the oscillator transistor is still
>>> cutoff for part of the oscillator cycle.
>>> The oscillator VCC is fed from a tap in a resistive divider network
>>> (located on the external (to the oven) buffer board) connected between
>>> the 5V supply and ground.
>>> Usually one just varies the emitter (or collector) dc current by
>>> selecting a resistor value.
>>>
>>> Bruce
>>>
>>>> *****************
>>>> ----- Original Message -----
>>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>>> Sent: Friday, July 03, 2009 8:21 PM
>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>
>>>>
>>>>
>>>>
>>>>> WarrenS wrote:
>>>>>
>>>>>
>>>>>> Here is my two cents worth
>>>>>> 20 mv output, sure sounds like something is broken.
>>>>>> It should be fixed before it is modified.
>>>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>>>>
>>>>>>
>>>>>>
>>>>> Not so, if the resistor values are correct this is the voltage that one
>>>>> should get.
>>>>> The question that remains: Is the voltage what the designer intended?
>>>>> With the Colpitts oscillator used limiting in the oscillator transistor
>>>>> (usually by current cutoff) is used to limit the amplitude.
>>>>> Just increasing the voltage without checking the resultant crystal
>>>>> current will be within permissable limits can be counter productive.
>>>>>
>>>>>
>>>>>
>>>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>>>> If you do 'need' to modify the gain,
>>>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>>>
>>>>>> ws
>>>>>>
>>>>>>
>>>>>>
>>>>> Bruce
>>>>>
>>>>>
>>>>>> ********************
>>>>>> ----- Original Message -----
>>>>>> From: "Mike Monett" <xde-l2g3@myamail.com>
>>>>>> To: <time-nuts@febo.com>
>>>>>> Sent: Friday, July 03, 2009 11:50 AM
>>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>> > An update on the Broken Ovenaire OSC 85-50.
>>>>>>>
>>>>>>> > I prepared a 'schematic' of the Output Board and the Oscillator
>>>>>>> > Board (attached) and I have lots of pictures of the external unit
>>>>>>> > and the insides if anyone is interested.
>>>>>>>
>>>>>>> > I resoldered all connections and replaced all transistors on the
>>>>>>> > Output Board and the Oscillator Board all to no benefit. I
>>>>>>> > measured all the components with an LCR meter and found the 0.01
>>>>>>> > uF bypass on the 330 ohm resistor in the emitter circuit of the
>>>>>>> > output transistor of the Output Board to be low and with a high
>>>>>>> > ESR. I replaced this with about a 20% increase in output amplitude
>>>>>>> > but still inadequate. I replaced the rest of the 0.01 uF caps on
>>>>>>> > the output board with no additional benefit. I transiently
>>>>>>> > disconnected the Red wires from the Oven Controller board and
>>>>>>> > there was no increase in output or significant increase in voltage
>>>>>>> > to the Oscillator Board.
>>>>>>>
>>>>>>> > Therefore, it appeared that a 'low output crystal' (if such a
>>>>>>> > thing exists) was the only logical explanation that I could come
>>>>>>> > up with. That seeming to be the case, there appeared to be only 4
>>>>>>> > options. 1. Toss the OCXO (sorry, too much effort so far). 2.
>>>>>>>
>>>>>>> > Build an external amplifier (seemingly too much additional
>>>>>>> > effort). 3. Try to adjust on the bias of the oscillator transistor
>>>>>>> > to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
>>>>>>> > value of the resistor in the emitter circuit of the Oscillator
>>>>>>> > Board to get more gain out of the last stage in the Oscillator
>>>>>>> > Board.
>>>>>>>
>>>>>>> > I replaced the 470 ohm resistor with a 47 ohm resistor and the
>>>>>>> > amplitude increased to about 0.4 V P-P into a 50 ohm load and was
>>>>>>> > sufficient to make it a usable OCXO again.
>>>>>>>
>>>>>>> > I reassembled, resealed with Epoxy and all seems well so far.
>>>>>>>
>>>>>>> > If anyone wants pictures or other info, please let me know.
>>>>>>>
>>>>>>> > Thanks for all the suggestions and help.
>>>>>>>
>>>>>>> > Joe
>>>>>>>
>>>>>>> Joe,
>>>>>>>
>>>>>>> Congratulations on getting your system to work!
>>>>>>>
>>>>>>> A couple of things. First, trying to measure the currents in the
>>>>>>> circuit with a ferrite toroid won't do you much good. You don't know
>>>>>>> what the currents should be, and the secondary of the toroid
>>>>>>> transformer requires a termination resistor. The value changes with
>>>>>>> the turns ratio.
>>>>>>>
>>>>>>> Just from looking at the circuit, the RF currents will be extremely
>>>>>>> low. This requires a large number of turns on the secondary, which
>>>>>>> will probably resonate at or below the 10MHz operating frequency due
>>>>>>> to stray capacitance from the connection to the scope. So it is
>>>>>>> unlikely you will get any useful progress in this direction.
>>>>>>>
>>>>>>> However, from the values on your schematic, the output tank circuit
>>>>>>> resonates at 9.602MHz with a Q of 9.6. So the tank is already well
>>>>>>> below resonance, which attenuates the output voltage.
>>>>>>>
>>>>>>> Any stray capacitance you add to the circuit will bring the resonant
>>>>>>> frequency lower, further aggravating the loss in signal.
>>>>>>>
>>>>>>> The output tank is tapped with the 75pF and 91pF in series. This
>>>>>>> further attenuates the signal.
>>>>>>>
>>>>>>> I'd change the circuit to a single capacitor across the tank with a
>>>>>>> small trim capacitor to tune it to resonance.
>>>>>>>
>>>>>>> To get the signal into 50 ohms for distribution, I'd add a limiter
>>>>>>> if you can tolerate a square wave output, or a good emitter follower
>>>>>>> if you need a sine wave. Take the output from the collector of the
>>>>>>> 2N2369 to get the maximum signal amplitude.
>>>>>>>
>>>>>>> Your original post mentions an output amplitude of 20mV. If the
>>>>>>> normal amplitude is around 2V, this represents a loss of 40dB. This
>>>>>>> is a huge loss in signal. The circuit obviously worked at one time,
>>>>>>> so there may well be some other hidden problem.
>>>>>>>
>>>>>>> It is possible the crystal is damaged, but this seems unlikely. A
>>>>>>> crystal oscillator probably won't even start if the signal level is
>>>>>>> down 40dB.
>>>>>>>
>>>>>>> You can check the oscillator and crystal in SPICE. Normally, the
>>>>>>> high Q of the crystal will make the analysis very slow. It could
>>>>>>> take many hours for the simulation to begin oscillating and
>>>>>>> stabilize at the final amplitude. The transient analysis requires a
>>>>>>> very fine time step for accuracy, and you could run out of memory
>>>>>>> before the simulation was complete.
>>>>>>>
>>>>>>> I have developed a much faster way of analyzing a crystal oscillator
>>>>>>> in SPICE. Instead of requiring tens or hundreds of thousands of
>>>>>>> simulated cycles, this method gives accurate results in only a few
>>>>>>> dozen cycles. For more information, please see "SPICE Analysis of
>>>>>>> Crystal Oscillators"
>>>>>>>
>>>>>>> http://pstca.com/spice/xtal/clapp.htm
>>>>>>>
>>>>>>> You can estimate the value of the crystal ESR by finding the Q of
>>>>>>> your crystal and working backwards.
>>>>>>>
>>>>>>> I'm attaching a gif of your schematic for reference. This is rotated
>>>>>>> 90 degrees and enhanced in LView Pro to improve the contrast.
>>>>>>>
>>>>>>> Please let me know if you have any questions.
>>>>>>>
>>>>>>> Thanks,
>>>>>>>
>>>>>>> Mike
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>> _______________________________________________
>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>> and follow the instructions there.
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>
>>>>>
>>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>>
>>>>
>>>
>>>
>>>
>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
>
>
>
>
BG
Bruce Griffiths
Sat, Jul 4, 2009 6:58 AM
Warren
The Vcc dc current doesn't change much if the circuit is oscillating (at
least it shouldn't if the circuit operating correctly - the active
devices aren't supposed to saturate as thats adds too much phase noise).
My simulations were over the 50-100C rang although temperature doesn't
affect the currents and voltages too much.
They also don't include loading of the frequency control pot wiper.
There is no major inconsistency other than the effect of the
unanticipated loading of the EFC pot wiper.
The simulated oscillator Vcc is actually about 3V when everything is
working properly.
NB its important to use models that accurately reflect the correct
transistor junction area just blindly tossing in 2N3904's increases the
current somewhat as the Vbe is lower for a given current than the actual
transistors. The simulated oscillator Vcc is then closer to 2.5V.
Earlier circuit schematics didn't include as many actual measured dc
voltages.
Bruce
WarrenS wrote:
Bruce
4ma? Interesting, Maybe IF the Osc was running, BUT
My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
With it NOT osc, which is almost the present case,
the Only Dc current drawn is.
1.75 V on 20K = 0.08ma (Q1 bias)
1.04 V on 2.2K = 0.47ma (Q1 collector)
0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
Osc board =1.32 ma
Buffer board
(2*4.7k) at 2.49 V = 0.26 ma
Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 11:08 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Warren
My Spice model of the dc levels in the oscillator indicate a current
more like 4mA in the 510 ohm resistor.
Bruce
WarrenS wrote:
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
A couple more thoughts;
Is the VCC (Red wire) current draw correct?.
Should be 1.3 ma (per the schematic values) with the 470 Ohm and
more like 7 plus ma with the 47 ohm.
AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
If both answers are YES, at least most things are working correctly.
As Bruce said:
"if the resistor values are correct this is the voltage(s) that one should get"
With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
AND the output amplitude is not yet being controlled by current cut off.
Something that needs to be asked is are all these the correct values.
It sounds like you got them by measuring and not from a "should be" schematic.
One last, way out thought.
It was my belief that very good osc have some sort of AGC to control the amplitude,
ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
ws
The Colpitts crystal oscillator limits the crystal current by being
periodically cutoff for part of the cycle.
There is an optimum duty cycle (usually something like 25%) that
corresponds to minimum phase noise.
The oscillator dc collector current determines the crystal current and
hence the output amplitude.
It is possible to perform AGC by varying the oscillator Vcc but the
output amplifier schematic appears to have no circuitry for this.
Some circuits actually do this, but the oscillator transistor is still
cutoff for part of the oscillator cycle.
The oscillator VCC is fed from a tap in a resistive divider network
(located on the external (to the oven) buffer board) connected between
the 5V supply and ground.
Usually one just varies the emitter (or collector) dc current by
selecting a resistor value.
Bruce
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:21 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
Not so, if the resistor values are correct this is the voltage that one
should get.
The question that remains: Is the voltage what the designer intended?
With the Colpitts oscillator used limiting in the oscillator transistor
(usually by current cutoff) is used to limit the amplitude.
Just increasing the voltage without checking the resultant crystal
current will be within permissable limits can be counter productive.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
An update on the Broken Ovenaire OSC 85-50.
I prepared a 'schematic' of the Output Board and the Oscillator
Board (attached) and I have lots of pictures of the external unit
and the insides if anyone is interested.
I resoldered all connections and replaced all transistors on the
Output Board and the Oscillator Board all to no benefit. I
measured all the components with an LCR meter and found the 0.01
uF bypass on the 330 ohm resistor in the emitter circuit of the
output transistor of the Output Board to be low and with a high
ESR. I replaced this with about a 20% increase in output amplitude
but still inadequate. I replaced the rest of the 0.01 uF caps on
the output board with no additional benefit. I transiently
disconnected the Red wires from the Oven Controller board and
there was no increase in output or significant increase in voltage
to the Oscillator Board.
Therefore, it appeared that a 'low output crystal' (if such a
thing exists) was the only logical explanation that I could come
up with. That seeming to be the case, there appeared to be only 4
options. 1. Toss the OCXO (sorry, too much effort so far). 2.
Build an external amplifier (seemingly too much additional
effort). 3. Try to adjust on the bias of the oscillator transistor
to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
value of the resistor in the emitter circuit of the Oscillator
Board to get more gain out of the last stage in the Oscillator
Board.
I replaced the 470 ohm resistor with a 47 ohm resistor and the
amplitude increased to about 0.4 V P-P into a 50 ohm load and was
sufficient to make it a usable OCXO again.
I reassembled, resealed with Epoxy and all seems well so far.
If anyone wants pictures or other info, please let me know.
Thanks for all the suggestions and help.
Joe,
Congratulations on getting your system to work!
A couple of things. First, trying to measure the currents in the
circuit with a ferrite toroid won't do you much good. You don't know
what the currents should be, and the secondary of the toroid
transformer requires a termination resistor. The value changes with
the turns ratio.
Just from looking at the circuit, the RF currents will be extremely
low. This requires a large number of turns on the secondary, which
will probably resonate at or below the 10MHz operating frequency due
to stray capacitance from the connection to the scope. So it is
unlikely you will get any useful progress in this direction.
However, from the values on your schematic, the output tank circuit
resonates at 9.602MHz with a Q of 9.6. So the tank is already well
below resonance, which attenuates the output voltage.
Any stray capacitance you add to the circuit will bring the resonant
frequency lower, further aggravating the loss in signal.
The output tank is tapped with the 75pF and 91pF in series. This
further attenuates the signal.
I'd change the circuit to a single capacitor across the tank with a
small trim capacitor to tune it to resonance.
To get the signal into 50 ohms for distribution, I'd add a limiter
if you can tolerate a square wave output, or a good emitter follower
if you need a sine wave. Take the output from the collector of the
2N2369 to get the maximum signal amplitude.
Your original post mentions an output amplitude of 20mV. If the
normal amplitude is around 2V, this represents a loss of 40dB. This
is a huge loss in signal. The circuit obviously worked at one time,
so there may well be some other hidden problem.
It is possible the crystal is damaged, but this seems unlikely. A
crystal oscillator probably won't even start if the signal level is
down 40dB.
You can check the oscillator and crystal in SPICE. Normally, the
high Q of the crystal will make the analysis very slow. It could
take many hours for the simulation to begin oscillating and
stabilize at the final amplitude. The transient analysis requires a
very fine time step for accuracy, and you could run out of memory
before the simulation was complete.
I have developed a much faster way of analyzing a crystal oscillator
in SPICE. Instead of requiring tens or hundreds of thousands of
simulated cycles, this method gives accurate results in only a few
dozen cycles. For more information, please see "SPICE Analysis of
Crystal Oscillators"
http://pstca.com/spice/xtal/clapp.htm
You can estimate the value of the crystal ESR by finding the Q of
your crystal and working backwards.
I'm attaching a gif of your schematic for reference. This is rotated
90 degrees and enhanced in LView Pro to improve the contrast.
Please let me know if you have any questions.
Thanks,
Mike
Warren
The Vcc dc current doesn't change much if the circuit is oscillating (at
least it shouldn't if the circuit operating correctly - the active
devices aren't supposed to saturate as thats adds too much phase noise).
My simulations were over the 50-100C rang although temperature doesn't
affect the currents and voltages too much.
They also don't include loading of the frequency control pot wiper.
There is no major inconsistency other than the effect of the
unanticipated loading of the EFC pot wiper.
The simulated oscillator Vcc is actually about 3V when everything is
working properly.
NB its important to use models that accurately reflect the correct
transistor junction area just blindly tossing in 2N3904's increases the
current somewhat as the Vbe is lower for a given current than the actual
transistors. The simulated oscillator Vcc is then closer to 2.5V.
Earlier circuit schematics didn't include as many actual measured dc
voltages.
Bruce
WarrenS wrote:
> Bruce
>
> 4ma? Interesting, Maybe IF the Osc was running, BUT
> My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
>
> With it NOT osc, which is almost the present case,
> the Only Dc current drawn is.
>
> 1.75 V on 20K = 0.08ma (Q1 bias)
> 1.04 V on 2.2K = 0.47ma (Q1 collector)
> 0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
> Osc board =1.32 ma
>
> Buffer board
> (2*4.7k) at 2.49 V = 0.26 ma
>
>
> Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
>
> ws
>
> **************
>
> ----- Original Message -----
> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
> Sent: Friday, July 03, 2009 11:08 PM
> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>
>
>
>> Warren
>>
>> My Spice model of the dc levels in the oscillator indicate a current
>> more like 4mA in the 510 ohm resistor.
>>
>> Bruce
>>
>> WarrenS wrote:
>>
>>> I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
>>>
>>> The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
>>> Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
>>> If I'm correct about that (this time) then, It has got to be one of three things.
>>> The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
>>> It should be easy enough to measure which one it is.
>>> The voltage at the junction of the two 4.7K resistors are WRONG.
>>> Looks like it is caused by an excess current draw on the Grn lead.
>>> Check the direction and leakage of the voltage controlled cap diode and bypass.
>>> If all else is right, Look for the cause of the extra loading current on the 510 ohm,
>>>
>>> WS
>>>
>>> ****************
>>>
>>> ----- Original Message -----
>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>> Sent: Friday, July 03, 2009 9:55 PM
>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>
>>>
>>>
>>>
>>>> WarrenS wrote:
>>>>
>>>>
>>>>> A couple more thoughts;
>>>>>
>>>>> Is the VCC (Red wire) current draw correct?.
>>>>> Should be 1.3 ma (per the schematic values) with the 470 Ohm and
>>>>> more like 7 plus ma with the 47 ohm.
>>>>>
>>>>> AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
>>>>> If both answers are YES, at least most things are working correctly.
>>>>>
>>>>> As Bruce said:
>>>>>
>>>>>
>>>>>
>>>>>> "if the resistor values are correct this is the voltage(s) that one should get"
>>>>>>
>>>>>>
>>>>>>
>>>>> With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
>>>>> BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
>>>>> AND the output amplitude is not yet being controlled by current cut off.
>>>>>
>>>>> Something that needs to be asked is are all these the correct values.
>>>>> It sounds like you got them by measuring and not from a "should be" schematic.
>>>>>
>>>>> One last, way out thought.
>>>>> It was my belief that very good osc have some sort of AGC to control the amplitude,
>>>>> ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
>>>>> If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
>>>>>
>>>>> ws
>>>>>
>>>>>
>>>>>
>>>>>
>>>> The Colpitts crystal oscillator limits the crystal current by being
>>>> periodically cutoff for part of the cycle.
>>>> There is an optimum duty cycle (usually something like 25%) that
>>>> corresponds to minimum phase noise.
>>>> The oscillator dc collector current determines the crystal current and
>>>> hence the output amplitude.
>>>> It is possible to perform AGC by varying the oscillator Vcc but the
>>>> output amplifier schematic appears to have no circuitry for this.
>>>> Some circuits actually do this, but the oscillator transistor is still
>>>> cutoff for part of the oscillator cycle.
>>>> The oscillator VCC is fed from a tap in a resistive divider network
>>>> (located on the external (to the oven) buffer board) connected between
>>>> the 5V supply and ground.
>>>> Usually one just varies the emitter (or collector) dc current by
>>>> selecting a resistor value.
>>>>
>>>> Bruce
>>>>
>>>>
>>>>> *****************
>>>>> ----- Original Message -----
>>>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>>>> Sent: Friday, July 03, 2009 8:21 PM
>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>> WarrenS wrote:
>>>>>>
>>>>>>
>>>>>>
>>>>>>> Here is my two cents worth
>>>>>>> 20 mv output, sure sounds like something is broken.
>>>>>>> It should be fixed before it is modified.
>>>>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>> Not so, if the resistor values are correct this is the voltage that one
>>>>>> should get.
>>>>>> The question that remains: Is the voltage what the designer intended?
>>>>>> With the Colpitts oscillator used limiting in the oscillator transistor
>>>>>> (usually by current cutoff) is used to limit the amplitude.
>>>>>> Just increasing the voltage without checking the resultant crystal
>>>>>> current will be within permissable limits can be counter productive.
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>>>>> If you do 'need' to modify the gain,
>>>>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>>>>
>>>>>>> ws
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>> Bruce
>>>>>>
>>>>>>
>>>>>>
>>>>>>> ********************
>>>>>>> ----- Original Message -----
>>>>>>> From: "Mike Monett" <xde-l2g3@myamail.com>
>>>>>>> To: <time-nuts@febo.com>
>>>>>>> Sent: Friday, July 03, 2009 11:50 AM
>>>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> > An update on the Broken Ovenaire OSC 85-50.
>>>>>>>>
>>>>>>>> > I prepared a 'schematic' of the Output Board and the Oscillator
>>>>>>>> > Board (attached) and I have lots of pictures of the external unit
>>>>>>>> > and the insides if anyone is interested.
>>>>>>>>
>>>>>>>> > I resoldered all connections and replaced all transistors on the
>>>>>>>> > Output Board and the Oscillator Board all to no benefit. I
>>>>>>>> > measured all the components with an LCR meter and found the 0.01
>>>>>>>> > uF bypass on the 330 ohm resistor in the emitter circuit of the
>>>>>>>> > output transistor of the Output Board to be low and with a high
>>>>>>>> > ESR. I replaced this with about a 20% increase in output amplitude
>>>>>>>> > but still inadequate. I replaced the rest of the 0.01 uF caps on
>>>>>>>> > the output board with no additional benefit. I transiently
>>>>>>>> > disconnected the Red wires from the Oven Controller board and
>>>>>>>> > there was no increase in output or significant increase in voltage
>>>>>>>> > to the Oscillator Board.
>>>>>>>>
>>>>>>>> > Therefore, it appeared that a 'low output crystal' (if such a
>>>>>>>> > thing exists) was the only logical explanation that I could come
>>>>>>>> > up with. That seeming to be the case, there appeared to be only 4
>>>>>>>> > options. 1. Toss the OCXO (sorry, too much effort so far). 2.
>>>>>>>>
>>>>>>>> > Build an external amplifier (seemingly too much additional
>>>>>>>> > effort). 3. Try to adjust on the bias of the oscillator transistor
>>>>>>>> > to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
>>>>>>>> > value of the resistor in the emitter circuit of the Oscillator
>>>>>>>> > Board to get more gain out of the last stage in the Oscillator
>>>>>>>> > Board.
>>>>>>>>
>>>>>>>> > I replaced the 470 ohm resistor with a 47 ohm resistor and the
>>>>>>>> > amplitude increased to about 0.4 V P-P into a 50 ohm load and was
>>>>>>>> > sufficient to make it a usable OCXO again.
>>>>>>>>
>>>>>>>> > I reassembled, resealed with Epoxy and all seems well so far.
>>>>>>>>
>>>>>>>> > If anyone wants pictures or other info, please let me know.
>>>>>>>>
>>>>>>>> > Thanks for all the suggestions and help.
>>>>>>>>
>>>>>>>> > Joe
>>>>>>>>
>>>>>>>> Joe,
>>>>>>>>
>>>>>>>> Congratulations on getting your system to work!
>>>>>>>>
>>>>>>>> A couple of things. First, trying to measure the currents in the
>>>>>>>> circuit with a ferrite toroid won't do you much good. You don't know
>>>>>>>> what the currents should be, and the secondary of the toroid
>>>>>>>> transformer requires a termination resistor. The value changes with
>>>>>>>> the turns ratio.
>>>>>>>>
>>>>>>>> Just from looking at the circuit, the RF currents will be extremely
>>>>>>>> low. This requires a large number of turns on the secondary, which
>>>>>>>> will probably resonate at or below the 10MHz operating frequency due
>>>>>>>> to stray capacitance from the connection to the scope. So it is
>>>>>>>> unlikely you will get any useful progress in this direction.
>>>>>>>>
>>>>>>>> However, from the values on your schematic, the output tank circuit
>>>>>>>> resonates at 9.602MHz with a Q of 9.6. So the tank is already well
>>>>>>>> below resonance, which attenuates the output voltage.
>>>>>>>>
>>>>>>>> Any stray capacitance you add to the circuit will bring the resonant
>>>>>>>> frequency lower, further aggravating the loss in signal.
>>>>>>>>
>>>>>>>> The output tank is tapped with the 75pF and 91pF in series. This
>>>>>>>> further attenuates the signal.
>>>>>>>>
>>>>>>>> I'd change the circuit to a single capacitor across the tank with a
>>>>>>>> small trim capacitor to tune it to resonance.
>>>>>>>>
>>>>>>>> To get the signal into 50 ohms for distribution, I'd add a limiter
>>>>>>>> if you can tolerate a square wave output, or a good emitter follower
>>>>>>>> if you need a sine wave. Take the output from the collector of the
>>>>>>>> 2N2369 to get the maximum signal amplitude.
>>>>>>>>
>>>>>>>> Your original post mentions an output amplitude of 20mV. If the
>>>>>>>> normal amplitude is around 2V, this represents a loss of 40dB. This
>>>>>>>> is a huge loss in signal. The circuit obviously worked at one time,
>>>>>>>> so there may well be some other hidden problem.
>>>>>>>>
>>>>>>>> It is possible the crystal is damaged, but this seems unlikely. A
>>>>>>>> crystal oscillator probably won't even start if the signal level is
>>>>>>>> down 40dB.
>>>>>>>>
>>>>>>>> You can check the oscillator and crystal in SPICE. Normally, the
>>>>>>>> high Q of the crystal will make the analysis very slow. It could
>>>>>>>> take many hours for the simulation to begin oscillating and
>>>>>>>> stabilize at the final amplitude. The transient analysis requires a
>>>>>>>> very fine time step for accuracy, and you could run out of memory
>>>>>>>> before the simulation was complete.
>>>>>>>>
>>>>>>>> I have developed a much faster way of analyzing a crystal oscillator
>>>>>>>> in SPICE. Instead of requiring tens or hundreds of thousands of
>>>>>>>> simulated cycles, this method gives accurate results in only a few
>>>>>>>> dozen cycles. For more information, please see "SPICE Analysis of
>>>>>>>> Crystal Oscillators"
>>>>>>>>
>>>>>>>> http://pstca.com/spice/xtal/clapp.htm
>>>>>>>>
>>>>>>>> You can estimate the value of the crystal ESR by finding the Q of
>>>>>>>> your crystal and working backwards.
>>>>>>>>
>>>>>>>> I'm attaching a gif of your schematic for reference. This is rotated
>>>>>>>> 90 degrees and enhanced in LView Pro to improve the contrast.
>>>>>>>>
>>>>>>>> Please let me know if you have any questions.
>>>>>>>>
>>>>>>>> Thanks,
>>>>>>>>
>>>>>>>> Mike
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>> _______________________________________________
>>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>> and follow the instructions there.
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>>
>>>>>
>>>>>
>>>>
>>>>
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>>
>>
>>
>>
>>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
W
WarrenS
Sat, Jul 4, 2009 7:24 AM
I get a 'if working correct' Vcc of about 3.5V
If the Vcc is 2.49 volts, I calculate a 3ma unaccounted for current draw from something.
So I have same conclusion, one or more of the following is wrong.
510 Ohm too high, 5 volts too low, un-shown current draw thru Leakey caps or extra circuits or bad data.
(plus the pot is (and should be) more likely a 10 K ohm)
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 11:58 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Warren
The Vcc dc current doesn't change much if the circuit is oscillating (at
least it shouldn't if the circuit operating correctly - the active
devices aren't supposed to saturate as thats adds too much phase noise).
My simulations were over the 50-100C rang although temperature doesn't
affect the currents and voltages too much.
They also don't include loading of the frequency control pot wiper.
There is no major inconsistency other than the effect of the
unanticipated loading of the EFC pot wiper.
The simulated oscillator Vcc is actually about 3V when everything is
working properly.
NB its important to use models that accurately reflect the correct
transistor junction area just blindly tossing in 2N3904's increases the
current somewhat as the Vbe is lower for a given current than the actual
transistors. The simulated oscillator Vcc is then closer to 2.5V.
Earlier circuit schematics didn't include as many actual measured dc
voltages.
Bruce
WarrenS wrote:
Bruce
4ma? Interesting, Maybe IF the Osc was running, BUT
My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
With it NOT osc, which is almost the present case,
the Only Dc current drawn is.
1.75 V on 20K = 0.08ma (Q1 bias)
1.04 V on 2.2K = 0.47ma (Q1 collector)
0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
Osc board =1.32 ma
Buffer board
(2*4.7k) at 2.49 V = 0.26 ma
Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 11:08 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Warren
My Spice model of the dc levels in the oscillator indicate a current
more like 4mA in the 510 ohm resistor.
Bruce
WarrenS wrote:
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
A couple more thoughts;
Is the VCC (Red wire) current draw correct?.
Should be 1.3 ma (per the schematic values) with the 470 Ohm and
more like 7 plus ma with the 47 ohm.
AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
If both answers are YES, at least most things are working correctly.
As Bruce said:
"if the resistor values are correct this is the voltage(s) that one should get"
With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
AND the output amplitude is not yet being controlled by current cut off.
Something that needs to be asked is are all these the correct values.
It sounds like you got them by measuring and not from a "should be" schematic.
One last, way out thought.
It was my belief that very good osc have some sort of AGC to control the amplitude,
ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
ws
The Colpitts crystal oscillator limits the crystal current by being
periodically cutoff for part of the cycle.
There is an optimum duty cycle (usually something like 25%) that
corresponds to minimum phase noise.
The oscillator dc collector current determines the crystal current and
hence the output amplitude.
It is possible to perform AGC by varying the oscillator Vcc but the
output amplifier schematic appears to have no circuitry for this.
Some circuits actually do this, but the oscillator transistor is still
cutoff for part of the oscillator cycle.
The oscillator VCC is fed from a tap in a resistive divider network
(located on the external (to the oven) buffer board) connected between
the 5V supply and ground.
Usually one just varies the emitter (or collector) dc current by
selecting a resistor value.
Bruce
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:21 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
Not so, if the resistor values are correct this is the voltage that one
should get.
The question that remains: Is the voltage what the designer intended?
With the Colpitts oscillator used limiting in the oscillator transistor
(usually by current cutoff) is used to limit the amplitude.
Just increasing the voltage without checking the resultant crystal
current will be within permissable limits can be counter productive.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
An update on the Broken Ovenaire OSC 85-50.
I prepared a 'schematic' of the Output Board and the Oscillator
Board (attached) and I have lots of pictures of the external unit
and the insides if anyone is interested.
I resoldered all connections and replaced all transistors on the
Output Board and the Oscillator Board all to no benefit. I
measured all the components with an LCR meter and found the 0.01
uF bypass on the 330 ohm resistor in the emitter circuit of the
output transistor of the Output Board to be low and with a high
ESR. I replaced this with about a 20% increase in output amplitude
but still inadequate. I replaced the rest of the 0.01 uF caps on
the output board with no additional benefit. I transiently
disconnected the Red wires from the Oven Controller board and
there was no increase in output or significant increase in voltage
to the Oscillator Board.
Therefore, it appeared that a 'low output crystal' (if such a
thing exists) was the only logical explanation that I could come
up with. That seeming to be the case, there appeared to be only 4
options. 1. Toss the OCXO (sorry, too much effort so far). 2.
Build an external amplifier (seemingly too much additional
effort). 3. Try to adjust on the bias of the oscillator transistor
to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
value of the resistor in the emitter circuit of the Oscillator
Board to get more gain out of the last stage in the Oscillator
Board.
I replaced the 470 ohm resistor with a 47 ohm resistor and the
amplitude increased to about 0.4 V P-P into a 50 ohm load and was
sufficient to make it a usable OCXO again.
I reassembled, resealed with Epoxy and all seems well so far.
If anyone wants pictures or other info, please let me know.
Thanks for all the suggestions and help.
Joe,
Congratulations on getting your system to work!
A couple of things. First, trying to measure the currents in the
circuit with a ferrite toroid won't do you much good. You don't know
what the currents should be, and the secondary of the toroid
transformer requires a termination resistor. The value changes with
the turns ratio.
Just from looking at the circuit, the RF currents will be extremely
low. This requires a large number of turns on the secondary, which
will probably resonate at or below the 10MHz operating frequency due
to stray capacitance from the connection to the scope. So it is
unlikely you will get any useful progress in this direction.
However, from the values on your schematic, the output tank circuit
resonates at 9.602MHz with a Q of 9.6. So the tank is already well
below resonance, which attenuates the output voltage.
Any stray capacitance you add to the circuit will bring the resonant
frequency lower, further aggravating the loss in signal.
The output tank is tapped with the 75pF and 91pF in series. This
further attenuates the signal.
I'd change the circuit to a single capacitor across the tank with a
small trim capacitor to tune it to resonance.
To get the signal into 50 ohms for distribution, I'd add a limiter
if you can tolerate a square wave output, or a good emitter follower
if you need a sine wave. Take the output from the collector of the
2N2369 to get the maximum signal amplitude.
Your original post mentions an output amplitude of 20mV. If the
normal amplitude is around 2V, this represents a loss of 40dB. This
is a huge loss in signal. The circuit obviously worked at one time,
so there may well be some other hidden problem.
It is possible the crystal is damaged, but this seems unlikely. A
crystal oscillator probably won't even start if the signal level is
down 40dB.
You can check the oscillator and crystal in SPICE. Normally, the
high Q of the crystal will make the analysis very slow. It could
take many hours for the simulation to begin oscillating and
stabilize at the final amplitude. The transient analysis requires a
very fine time step for accuracy, and you could run out of memory
before the simulation was complete.
I have developed a much faster way of analyzing a crystal oscillator
in SPICE. Instead of requiring tens or hundreds of thousands of
simulated cycles, this method gives accurate results in only a few
dozen cycles. For more information, please see "SPICE Analysis of
Crystal Oscillators"
http://pstca.com/spice/xtal/clapp.htm
You can estimate the value of the crystal ESR by finding the Q of
your crystal and working backwards.
I'm attaching a gif of your schematic for reference. This is rotated
90 degrees and enhanced in LView Pro to improve the contrast.
Please let me know if you have any questions.
Thanks,
Mike
I get a 'if working correct' Vcc of about 3.5V
If the Vcc is 2.49 volts, I calculate a 3ma unaccounted for current draw from something.
So I have same conclusion, one or more of the following is wrong.
510 Ohm too high, 5 volts too low, un-shown current draw thru Leakey caps or extra circuits or bad data.
(plus the pot is (and should be) more likely a 10 K ohm)
ws
******************
----- Original Message -----
From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
Sent: Friday, July 03, 2009 11:58 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
> Warren
>
> The Vcc dc current doesn't change much if the circuit is oscillating (at
> least it shouldn't if the circuit operating correctly - the active
> devices aren't supposed to saturate as thats adds too much phase noise).
>
> My simulations were over the 50-100C rang although temperature doesn't
> affect the currents and voltages too much.
> They also don't include loading of the frequency control pot wiper.
>
> There is no major inconsistency other than the effect of the
> unanticipated loading of the EFC pot wiper.
> The simulated oscillator Vcc is actually about 3V when everything is
> working properly.
> NB its important to use models that accurately reflect the correct
> transistor junction area just blindly tossing in 2N3904's increases the
> current somewhat as the Vbe is lower for a given current than the actual
> transistors. The simulated oscillator Vcc is then closer to 2.5V.
>
> Earlier circuit schematics didn't include as many actual measured dc
> voltages.
>
> Bruce
>
> WarrenS wrote:
>> Bruce
>>
>> 4ma? Interesting, Maybe IF the Osc was running, BUT
>> My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
>>
>> With it NOT osc, which is almost the present case,
>> the Only Dc current drawn is.
>>
>> 1.75 V on 20K = 0.08ma (Q1 bias)
>> 1.04 V on 2.2K = 0.47ma (Q1 collector)
>> 0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
>> Osc board =1.32 ma
>>
>> Buffer board
>> (2*4.7k) at 2.49 V = 0.26 ma
>>
>>
>> Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
>>
>> ws
>>
>> **************
>>
>> ----- Original Message -----
>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>> Sent: Friday, July 03, 2009 11:08 PM
>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>
>>
>>
>>> Warren
>>>
>>> My Spice model of the dc levels in the oscillator indicate a current
>>> more like 4mA in the 510 ohm resistor.
>>>
>>> Bruce
>>>
>>> WarrenS wrote:
>>>
>>>> I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
>>>>
>>>> The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
>>>> Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
>>>> If I'm correct about that (this time) then, It has got to be one of three things.
>>>> The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
>>>> It should be easy enough to measure which one it is.
>>>> The voltage at the junction of the two 4.7K resistors are WRONG.
>>>> Looks like it is caused by an excess current draw on the Grn lead.
>>>> Check the direction and leakage of the voltage controlled cap diode and bypass.
>>>> If all else is right, Look for the cause of the extra loading current on the 510 ohm,
>>>>
>>>> WS
>>>>
>>>> ****************
>>>>
>>>> ----- Original Message -----
>>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>>> Sent: Friday, July 03, 2009 9:55 PM
>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>
>>>>
>>>>
>>>>
>>>>> WarrenS wrote:
>>>>>
>>>>>
>>>>>> A couple more thoughts;
>>>>>>
>>>>>> Is the VCC (Red wire) current draw correct?.
>>>>>> Should be 1.3 ma (per the schematic values) with the 470 Ohm and
>>>>>> more like 7 plus ma with the 47 ohm.
>>>>>>
>>>>>> AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
>>>>>> If both answers are YES, at least most things are working correctly.
>>>>>>
>>>>>> As Bruce said:
>>>>>>
>>>>>>
>>>>>>
>>>>>>> "if the resistor values are correct this is the voltage(s) that one should get"
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>> With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
>>>>>> BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
>>>>>> AND the output amplitude is not yet being controlled by current cut off.
>>>>>>
>>>>>> Something that needs to be asked is are all these the correct values.
>>>>>> It sounds like you got them by measuring and not from a "should be" schematic.
>>>>>>
>>>>>> One last, way out thought.
>>>>>> It was my belief that very good osc have some sort of AGC to control the amplitude,
>>>>>> ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
>>>>>> If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
>>>>>>
>>>>>> ws
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>> The Colpitts crystal oscillator limits the crystal current by being
>>>>> periodically cutoff for part of the cycle.
>>>>> There is an optimum duty cycle (usually something like 25%) that
>>>>> corresponds to minimum phase noise.
>>>>> The oscillator dc collector current determines the crystal current and
>>>>> hence the output amplitude.
>>>>> It is possible to perform AGC by varying the oscillator Vcc but the
>>>>> output amplifier schematic appears to have no circuitry for this.
>>>>> Some circuits actually do this, but the oscillator transistor is still
>>>>> cutoff for part of the oscillator cycle.
>>>>> The oscillator VCC is fed from a tap in a resistive divider network
>>>>> (located on the external (to the oven) buffer board) connected between
>>>>> the 5V supply and ground.
>>>>> Usually one just varies the emitter (or collector) dc current by
>>>>> selecting a resistor value.
>>>>>
>>>>> Bruce
>>>>>
>>>>>
>>>>>> *****************
>>>>>> ----- Original Message -----
>>>>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>>>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>>>>> Sent: Friday, July 03, 2009 8:21 PM
>>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>> WarrenS wrote:
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> Here is my two cents worth
>>>>>>>> 20 mv output, sure sounds like something is broken.
>>>>>>>> It should be fixed before it is modified.
>>>>>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>> Not so, if the resistor values are correct this is the voltage that one
>>>>>>> should get.
>>>>>>> The question that remains: Is the voltage what the designer intended?
>>>>>>> With the Colpitts oscillator used limiting in the oscillator transistor
>>>>>>> (usually by current cutoff) is used to limit the amplitude.
>>>>>>> Just increasing the voltage without checking the resultant crystal
>>>>>>> current will be within permissable limits can be counter productive.
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>>>>>> If you do 'need' to modify the gain,
>>>>>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>>>>>
>>>>>>>> ws
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>> Bruce
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> ********************
>>>>>>>> ----- Original Message -----
>>>>>>>> From: "Mike Monett" <xde-l2g3@myamail.com>
>>>>>>>> To: <time-nuts@febo.com>
>>>>>>>> Sent: Friday, July 03, 2009 11:50 AM
>>>>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>> > An update on the Broken Ovenaire OSC 85-50.
>>>>>>>>>
>>>>>>>>> > I prepared a 'schematic' of the Output Board and the Oscillator
>>>>>>>>> > Board (attached) and I have lots of pictures of the external unit
>>>>>>>>> > and the insides if anyone is interested.
>>>>>>>>>
>>>>>>>>> > I resoldered all connections and replaced all transistors on the
>>>>>>>>> > Output Board and the Oscillator Board all to no benefit. I
>>>>>>>>> > measured all the components with an LCR meter and found the 0.01
>>>>>>>>> > uF bypass on the 330 ohm resistor in the emitter circuit of the
>>>>>>>>> > output transistor of the Output Board to be low and with a high
>>>>>>>>> > ESR. I replaced this with about a 20% increase in output amplitude
>>>>>>>>> > but still inadequate. I replaced the rest of the 0.01 uF caps on
>>>>>>>>> > the output board with no additional benefit. I transiently
>>>>>>>>> > disconnected the Red wires from the Oven Controller board and
>>>>>>>>> > there was no increase in output or significant increase in voltage
>>>>>>>>> > to the Oscillator Board.
>>>>>>>>>
>>>>>>>>> > Therefore, it appeared that a 'low output crystal' (if such a
>>>>>>>>> > thing exists) was the only logical explanation that I could come
>>>>>>>>> > up with. That seeming to be the case, there appeared to be only 4
>>>>>>>>> > options. 1. Toss the OCXO (sorry, too much effort so far). 2.
>>>>>>>>>
>>>>>>>>> > Build an external amplifier (seemingly too much additional
>>>>>>>>> > effort). 3. Try to adjust on the bias of the oscillator transistor
>>>>>>>>> > to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
>>>>>>>>> > value of the resistor in the emitter circuit of the Oscillator
>>>>>>>>> > Board to get more gain out of the last stage in the Oscillator
>>>>>>>>> > Board.
>>>>>>>>>
>>>>>>>>> > I replaced the 470 ohm resistor with a 47 ohm resistor and the
>>>>>>>>> > amplitude increased to about 0.4 V P-P into a 50 ohm load and was
>>>>>>>>> > sufficient to make it a usable OCXO again.
>>>>>>>>>
>>>>>>>>> > I reassembled, resealed with Epoxy and all seems well so far.
>>>>>>>>>
>>>>>>>>> > If anyone wants pictures or other info, please let me know.
>>>>>>>>>
>>>>>>>>> > Thanks for all the suggestions and help.
>>>>>>>>>
>>>>>>>>> > Joe
>>>>>>>>>
>>>>>>>>> Joe,
>>>>>>>>>
>>>>>>>>> Congratulations on getting your system to work!
>>>>>>>>>
>>>>>>>>> A couple of things. First, trying to measure the currents in the
>>>>>>>>> circuit with a ferrite toroid won't do you much good. You don't know
>>>>>>>>> what the currents should be, and the secondary of the toroid
>>>>>>>>> transformer requires a termination resistor. The value changes with
>>>>>>>>> the turns ratio.
>>>>>>>>>
>>>>>>>>> Just from looking at the circuit, the RF currents will be extremely
>>>>>>>>> low. This requires a large number of turns on the secondary, which
>>>>>>>>> will probably resonate at or below the 10MHz operating frequency due
>>>>>>>>> to stray capacitance from the connection to the scope. So it is
>>>>>>>>> unlikely you will get any useful progress in this direction.
>>>>>>>>>
>>>>>>>>> However, from the values on your schematic, the output tank circuit
>>>>>>>>> resonates at 9.602MHz with a Q of 9.6. So the tank is already well
>>>>>>>>> below resonance, which attenuates the output voltage.
>>>>>>>>>
>>>>>>>>> Any stray capacitance you add to the circuit will bring the resonant
>>>>>>>>> frequency lower, further aggravating the loss in signal.
>>>>>>>>>
>>>>>>>>> The output tank is tapped with the 75pF and 91pF in series. This
>>>>>>>>> further attenuates the signal.
>>>>>>>>>
>>>>>>>>> I'd change the circuit to a single capacitor across the tank with a
>>>>>>>>> small trim capacitor to tune it to resonance.
>>>>>>>>>
>>>>>>>>> To get the signal into 50 ohms for distribution, I'd add a limiter
>>>>>>>>> if you can tolerate a square wave output, or a good emitter follower
>>>>>>>>> if you need a sine wave. Take the output from the collector of the
>>>>>>>>> 2N2369 to get the maximum signal amplitude.
>>>>>>>>>
>>>>>>>>> Your original post mentions an output amplitude of 20mV. If the
>>>>>>>>> normal amplitude is around 2V, this represents a loss of 40dB. This
>>>>>>>>> is a huge loss in signal. The circuit obviously worked at one time,
>>>>>>>>> so there may well be some other hidden problem.
>>>>>>>>>
>>>>>>>>> It is possible the crystal is damaged, but this seems unlikely. A
>>>>>>>>> crystal oscillator probably won't even start if the signal level is
>>>>>>>>> down 40dB.
>>>>>>>>>
>>>>>>>>> You can check the oscillator and crystal in SPICE. Normally, the
>>>>>>>>> high Q of the crystal will make the analysis very slow. It could
>>>>>>>>> take many hours for the simulation to begin oscillating and
>>>>>>>>> stabilize at the final amplitude. The transient analysis requires a
>>>>>>>>> very fine time step for accuracy, and you could run out of memory
>>>>>>>>> before the simulation was complete.
>>>>>>>>>
>>>>>>>>> I have developed a much faster way of analyzing a crystal oscillator
>>>>>>>>> in SPICE. Instead of requiring tens or hundreds of thousands of
>>>>>>>>> simulated cycles, this method gives accurate results in only a few
>>>>>>>>> dozen cycles. For more information, please see "SPICE Analysis of
>>>>>>>>> Crystal Oscillators"
>>>>>>>>>
>>>>>>>>> http://pstca.com/spice/xtal/clapp.htm
>>>>>>>>>
>>>>>>>>> You can estimate the value of the crystal ESR by finding the Q of
>>>>>>>>> your crystal and working backwards.
>>>>>>>>>
>>>>>>>>> I'm attaching a gif of your schematic for reference. This is rotated
>>>>>>>>> 90 degrees and enhanced in LView Pro to improve the contrast.
>>>>>>>>>
>>>>>>>>> Please let me know if you have any questions.
>>>>>>>>>
>>>>>>>>> Thanks,
>>>>>>>>>
>>>>>>>>> Mike
>>>>>>>>>
>>>>>>>>>
>>>>>>>>>
>>>>>>>>>
>>>>>>>>>
>>>>>>>> _______________________________________________
>>>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>>> and follow the instructions there.
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>> _______________________________________________
>>>>>> time-nuts mailing list -- time-nuts@febo.com
>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>> and follow the instructions there.
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>
>>>>>
>>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>>
>>>>
>>>
>>>
>>>
>>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
>
>
>
>
KS
Kit Scally
Sat, Jul 4, 2009 8:21 AM
G'day Joe,
My call on <this> question is: "very seldom with a properly designed
circuit and crystal" (NO flames please on these descriptions !)
Having said that, I have a couple of broken crystals from once-working
HP E1938A oscillators. All appear to have some form of physical
fracture or electrical disconnect within the copper crystal capsule. A
careful shake makes a re-connection but fails if the capsule is moved
even slightly.
I've not attempted to butcher a capsule open to determine the cause -
this to my mind is bordering on the sacrilegious but if given absolution
from another TimeNut, I may attempt it and photograph the results ...
Regards,
Kit
VK2LL
Do XTAL's fail often? If so, by what mechanism? Slowly fade away?
Sudden death? I am not experienced enough to know this answer but in
all my years with various HAM gear, I do not recall the failure of a
single XTAL. However, none of these were in an oven.
Thanks again for all the useful information.
Joe
G'day Joe,
My call on <this> question is: "very seldom with a properly designed
circuit and crystal" (NO flames please on these descriptions !)
Having said that, I have a couple of broken crystals from once-working
HP E1938A oscillators. All appear to have some form of physical
fracture or electrical disconnect within the copper crystal capsule. A
careful shake makes a re-connection but fails if the capsule is moved
even slightly.
I've not attempted to butcher a capsule open to determine the cause -
this to my mind is bordering on the sacrilegious but if given absolution
from another TimeNut, I may attempt it and photograph the results ...
Regards,
Kit
VK2LL
*************
Do XTAL's fail often? If so, by what mechanism? Slowly fade away?
Sudden death? I am not experienced enough to know this answer but in
all my years with various HAM gear, I do not recall the failure of a
single XTAL. However, none of these were in an oven.
Thanks again for all the useful information.
Joe
BG
Bruce Griffiths
Sat, Jul 4, 2009 8:38 AM
Kit
If you are confident of your diagnosis then you have little to lose.
The crystal is defunct and isn't going to fix itself so it may be
instructive to take a peak inside.
However don't the E1938 cans have an internal vacuum?
Some care in opening may be required.
Bruce
Kit Scally wrote:
G'day Joe,
My call on <this> question is: "very seldom with a properly designed
circuit and crystal" (NO flames please on these descriptions !)
Having said that, I have a couple of broken crystals from once-working
HP E1938A oscillators. All appear to have some form of physical
fracture or electrical disconnect within the copper crystal capsule. A
careful shake makes a re-connection but fails if the capsule is moved
even slightly.
I've not attempted to butcher a capsule open to determine the cause -
this to my mind is bordering on the sacrilegious but if given absolution
from another TimeNut, I may attempt it and photograph the results ...
Regards,
Kit
VK2LL
Do XTAL's fail often? If so, by what mechanism? Slowly fade away?
Sudden death? I am not experienced enough to know this answer but in
all my years with various HAM gear, I do not recall the failure of a
single XTAL. However, none of these were in an oven.
Thanks again for all the useful information.
Joe
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Kit
If you are confident of your diagnosis then you have little to lose.
The crystal is defunct and isn't going to fix itself so it may be
instructive to take a peak inside.
However don't the E1938 cans have an internal vacuum?
Some care in opening may be required.
Bruce
Kit Scally wrote:
> G'day Joe,
>
>
> My call on <this> question is: "very seldom with a properly designed
> circuit and crystal" (NO flames please on these descriptions !)
> Having said that, I have a couple of broken crystals from once-working
> HP E1938A oscillators. All appear to have some form of physical
> fracture or electrical disconnect within the copper crystal capsule. A
> careful shake makes a re-connection but fails if the capsule is moved
> even slightly.
> I've not attempted to butcher a capsule open to determine the cause -
> this to my mind is bordering on the sacrilegious but if given absolution
> from another TimeNut, I may attempt it and photograph the results ...
>
> Regards,
>
>
> Kit
> VK2LL
>
> *************
> Do XTAL's fail often? If so, by what mechanism? Slowly fade away?
> Sudden death? I am not experienced enough to know this answer but in
> all my years with various HAM gear, I do not recall the failure of a
> single XTAL. However, none of these were in an oven.
>
> Thanks again for all the useful information.
>
> Joe
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
KS
Kit Scally
Sat, Jul 4, 2009 8:56 AM
Bruce,
I'm 99% confident - I had the whole hockey-puk apart and manipulated the
capsule to get the unit working. The assembly isn't under vacuum - it
just requires a bit of "knack" to prise it apart. However, the crystal
puk is, I'm sure, under vacuum or filled with an inert gas (comments
please, Rick ?). It's about 15mm diam. I have some pix of the puk.
Can you email me privately how to upload pix to "somewhere" - I'll
download them tomorrow.
I'll wear safetly glasses ...
Kit
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Bruce Griffiths
Sent: Saturday, 4 July 2009 6:38 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50 - crystal failure ?
Kit
If you are confident of your diagnosis then you have little to lose. The
crystal is defunct and isn't going to fix itself so it may be
instructive to take a peak inside. However don't the E1938 cans have an
internal vacuum? Some care in opening may be required.
Bruce
Kit Scally wrote:
G'day Joe,
My call on <this> question is: "very seldom with a properly designed
circuit and crystal" (NO flames please on these descriptions !) Having
said that, I have a couple of broken crystals from once-working HP
E1938A oscillators. All appear to have some form of physical fracture
or electrical disconnect within the copper crystal capsule. A careful
shake makes a re-connection but fails if the capsule is moved even
slightly. I've not attempted to butcher a capsule open to determine
the cause - this to my mind is bordering on the sacrilegious but if
given absolution from another TimeNut, I may attempt it and photograph
the results ...
Regards,
Kit
VK2LL
Do XTAL's fail often? If so, by what mechanism? Slowly fade away?
Sudden death? I am not experienced enough to know this answer but in
all my years with various HAM gear, I do not recall the failure of a
single XTAL. However, none of these were in an oven.
Thanks again for all the useful information.
Joe
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Bruce,
I'm 99% confident - I had the whole hockey-puk apart and manipulated the
capsule to get the unit working. The assembly isn't under vacuum - it
just requires a bit of "knack" to prise it apart. However, the crystal
puk is, I'm sure, under vacuum or filled with an inert gas (comments
please, Rick ?). It's about 15mm diam. I have some pix of the puk.
Can you email me privately how to upload pix to "somewhere" - I'll
download them tomorrow.
I'll wear safetly glasses ...
Kit
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Bruce Griffiths
Sent: Saturday, 4 July 2009 6:38 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50 - crystal failure ?
Kit
If you are confident of your diagnosis then you have little to lose. The
crystal is defunct and isn't going to fix itself so it may be
instructive to take a peak inside. However don't the E1938 cans have an
internal vacuum? Some care in opening may be required.
Bruce
Kit Scally wrote:
> G'day Joe,
>
>
> My call on <this> question is: "very seldom with a properly designed
> circuit and crystal" (NO flames please on these descriptions !) Having
> said that, I have a couple of broken crystals from once-working HP
> E1938A oscillators. All appear to have some form of physical fracture
> or electrical disconnect within the copper crystal capsule. A careful
> shake makes a re-connection but fails if the capsule is moved even
> slightly. I've not attempted to butcher a capsule open to determine
> the cause - this to my mind is bordering on the sacrilegious but if
> given absolution from another TimeNut, I may attempt it and photograph
> the results ...
>
> Regards,
>
>
> Kit
> VK2LL
>
> *************
> Do XTAL's fail often? If so, by what mechanism? Slowly fade away?
> Sudden death? I am not experienced enough to know this answer but in
> all my years with various HAM gear, I do not recall the failure of a
> single XTAL. However, none of these were in an oven.
>
> Thanks again for all the useful information.
>
> Joe
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
W
WarrenS
Sat, Jul 4, 2009 10:05 AM
Here is a new observation about low Vcc.
It is not likely that the Osc Vcc voltage and the Varicap freq control voltage is coming from the same 510 Ohm resistor without a regulator.
Something is missing on the schematic, because now any Osc current change would cause the freq to change.
And there is the missing 3 plus ma that can not be accounted for in your voltage measurements.
The Osc and pot circuits as shown are drawing about 1.5 ma, and there seems to be almost 5ma going thru the 510 ohm.
My guess is that this is from a 2.50 volt precision shunt regulator on the Vcc, in the Osc oven, but not shown on the schematic.
That should be pretty easy to find.
If so, make sure your 47 ohm fix does not cause the 2.490 volts to loose its regulation when the circuit is put back in the oven.
The increase temp will cause the osc current to increase.
If the EFC pot is really 10K ohm end to end, that would account for the other strange voltage readings.
Also if the Freq as small as it is, is still correct, that would eliminate most of the bad xtral possibilities.
ws
----- Original Message -----
From: "WarrenS" warrensjmail-one@yahoo.com
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Saturday, July 04, 2009 12:24 AM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
I get a 'if working correct' Vcc of about 3.5V
If the Vcc is 2.49 volts, I calculate a 3ma unaccounted for current draw from something.
So I have same conclusion, one or more of the following is wrong.
510 Ohm too high, 5 volts too low, un-shown current draw thru Leakey caps or extra circuits or bad data.
(plus the pot is (and should be) more likely a 10 K ohm)
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 11:58 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Warren
The Vcc dc current doesn't change much if the circuit is oscillating (at
least it shouldn't if the circuit operating correctly - the active
devices aren't supposed to saturate as thats adds too much phase noise).
My simulations were over the 50-100C rang although temperature doesn't
affect the currents and voltages too much.
They also don't include loading of the frequency control pot wiper.
There is no major inconsistency other than the effect of the
unanticipated loading of the EFC pot wiper.
The simulated oscillator Vcc is actually about 3V when everything is
working properly.
NB its important to use models that accurately reflect the correct
transistor junction area just blindly tossing in 2N3904's increases the
current somewhat as the Vbe is lower for a given current than the actual
transistors. The simulated oscillator Vcc is then closer to 2.5V.
Earlier circuit schematics didn't include as many actual measured dc
voltages.
Bruce
WarrenS wrote:
Bruce
4ma? Interesting, Maybe IF the Osc was running, BUT
My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
With it NOT osc, which is almost the present case,
the Only Dc current drawn is.
1.75 V on 20K = 0.08ma (Q1 bias)
1.04 V on 2.2K = 0.47ma (Q1 collector)
0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
Osc board =1.32 ma
Buffer board
(2*4.7k) at 2.49 V = 0.26 ma
Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
ws
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 11:08 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Warren
My Spice model of the dc levels in the oscillator indicate a current
more like 4mA in the 510 ohm resistor.
Bruce
WarrenS wrote:
I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
If I'm correct about that (this time) then, It has got to be one of three things.
The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
It should be easy enough to measure which one it is.
The voltage at the junction of the two 4.7K resistors are WRONG.
Looks like it is caused by an excess current draw on the Grn lead.
Check the direction and leakage of the voltage controlled cap diode and bypass.
If all else is right, Look for the cause of the extra loading current on the 510 ohm,
WS
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 9:55 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
A couple more thoughts;
Is the VCC (Red wire) current draw correct?.
Should be 1.3 ma (per the schematic values) with the 470 Ohm and
more like 7 plus ma with the 47 ohm.
AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
If both answers are YES, at least most things are working correctly.
As Bruce said:
"if the resistor values are correct this is the voltage(s) that one should get"
With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
AND the output amplitude is not yet being controlled by current cut off.
Something that needs to be asked is are all these the correct values.
It sounds like you got them by measuring and not from a "should be" schematic.
One last, way out thought.
It was my belief that very good osc have some sort of AGC to control the amplitude,
ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
ws
The Colpitts crystal oscillator limits the crystal current by being
periodically cutoff for part of the cycle.
There is an optimum duty cycle (usually something like 25%) that
corresponds to minimum phase noise.
The oscillator dc collector current determines the crystal current and
hence the output amplitude.
It is possible to perform AGC by varying the oscillator Vcc but the
output amplifier schematic appears to have no circuitry for this.
Some circuits actually do this, but the oscillator transistor is still
cutoff for part of the oscillator cycle.
The oscillator VCC is fed from a tap in a resistive divider network
(located on the external (to the oven) buffer board) connected between
the 5V supply and ground.
Usually one just varies the emitter (or collector) dc current by
selecting a resistor value.
Bruce
----- Original Message -----
From: "Bruce Griffiths" bruce.griffiths@xtra.co.nz
To: "Discussion of precise time and frequency measurement" time-nuts@febo.com
Sent: Friday, July 03, 2009 8:21 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
Here is my two cents worth
20 mv output, sure sounds like something is broken.
It should be fixed before it is modified.
The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
Not so, if the resistor values are correct this is the voltage that one
should get.
The question that remains: Is the voltage what the designer intended?
With the Colpitts oscillator used limiting in the oscillator transistor
(usually by current cutoff) is used to limit the amplitude.
Just increasing the voltage without checking the resultant crystal
current will be within permissable limits can be counter productive.
The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
If you do 'need' to modify the gain,
It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
ws
An update on the Broken Ovenaire OSC 85-50.
I prepared a 'schematic' of the Output Board and the Oscillator
Board (attached) and I have lots of pictures of the external unit
and the insides if anyone is interested.
I resoldered all connections and replaced all transistors on the
Output Board and the Oscillator Board all to no benefit. I
measured all the components with an LCR meter and found the 0.01
uF bypass on the 330 ohm resistor in the emitter circuit of the
output transistor of the Output Board to be low and with a high
ESR. I replaced this with about a 20% increase in output amplitude
but still inadequate. I replaced the rest of the 0.01 uF caps on
the output board with no additional benefit. I transiently
disconnected the Red wires from the Oven Controller board and
there was no increase in output or significant increase in voltage
to the Oscillator Board.
Therefore, it appeared that a 'low output crystal' (if such a
thing exists) was the only logical explanation that I could come
up with. That seeming to be the case, there appeared to be only 4
options. 1. Toss the OCXO (sorry, too much effort so far). 2.
Build an external amplifier (seemingly too much additional
effort). 3. Try to adjust on the bias of the oscillator transistor
to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
value of the resistor in the emitter circuit of the Oscillator
Board to get more gain out of the last stage in the Oscillator
Board.
I replaced the 470 ohm resistor with a 47 ohm resistor and the
amplitude increased to about 0.4 V P-P into a 50 ohm load and was
sufficient to make it a usable OCXO again.
I reassembled, resealed with Epoxy and all seems well so far.
If anyone wants pictures or other info, please let me know.
Thanks for all the suggestions and help.
Joe,
Congratulations on getting your system to work!
A couple of things. First, trying to measure the currents in the
circuit with a ferrite toroid won't do you much good. You don't know
what the currents should be, and the secondary of the toroid
transformer requires a termination resistor. The value changes with
the turns ratio.
Just from looking at the circuit, the RF currents will be extremely
low. This requires a large number of turns on the secondary, which
will probably resonate at or below the 10MHz operating frequency due
to stray capacitance from the connection to the scope. So it is
unlikely you will get any useful progress in this direction.
However, from the values on your schematic, the output tank circuit
resonates at 9.602MHz with a Q of 9.6. So the tank is already well
below resonance, which attenuates the output voltage.
Any stray capacitance you add to the circuit will bring the resonant
frequency lower, further aggravating the loss in signal.
The output tank is tapped with the 75pF and 91pF in series. This
further attenuates the signal.
I'd change the circuit to a single capacitor across the tank with a
small trim capacitor to tune it to resonance.
To get the signal into 50 ohms for distribution, I'd add a limiter
if you can tolerate a square wave output, or a good emitter follower
if you need a sine wave. Take the output from the collector of the
2N2369 to get the maximum signal amplitude.
Your original post mentions an output amplitude of 20mV. If the
normal amplitude is around 2V, this represents a loss of 40dB. This
is a huge loss in signal. The circuit obviously worked at one time,
so there may well be some other hidden problem.
It is possible the crystal is damaged, but this seems unlikely. A
crystal oscillator probably won't even start if the signal level is
down 40dB.
You can check the oscillator and crystal in SPICE. Normally, the
high Q of the crystal will make the analysis very slow. It could
take many hours for the simulation to begin oscillating and
stabilize at the final amplitude. The transient analysis requires a
very fine time step for accuracy, and you could run out of memory
before the simulation was complete.
I have developed a much faster way of analyzing a crystal oscillator
in SPICE. Instead of requiring tens or hundreds of thousands of
simulated cycles, this method gives accurate results in only a few
dozen cycles. For more information, please see "SPICE Analysis of
Crystal Oscillators"
http://pstca.com/spice/xtal/clapp.htm
You can estimate the value of the crystal ESR by finding the Q of
your crystal and working backwards.
I'm attaching a gif of your schematic for reference. This is rotated
90 degrees and enhanced in LView Pro to improve the contrast.
Please let me know if you have any questions.
Thanks,
Mike
Here is a new observation about low Vcc.
It is not likely that the Osc Vcc voltage and the Varicap freq control voltage is coming from the same 510 Ohm resistor without a regulator.
Something is missing on the schematic, because now any Osc current change would cause the freq to change.
And there is the missing 3 plus ma that can not be accounted for in your voltage measurements.
The Osc and pot circuits as shown are drawing about 1.5 ma, and there seems to be almost 5ma going thru the 510 ohm.
My guess is that this is from a 2.50 volt precision shunt regulator on the Vcc, in the Osc oven, but not shown on the schematic.
That should be pretty easy to find.
If so, make sure your 47 ohm fix does not cause the 2.490 volts to loose its regulation when the circuit is put back in the oven.
The increase temp will cause the osc current to increase.
If the EFC pot is really 10K ohm end to end, that would account for the other strange voltage readings.
Also if the Freq as small as it is, is still correct, that would eliminate most of the bad xtral possibilities.
ws
*****************
----- Original Message -----
From: "WarrenS" <warrensjmail-one@yahoo.com>
To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
Sent: Saturday, July 04, 2009 12:24 AM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
I get a 'if working correct' Vcc of about 3.5V
If the Vcc is 2.49 volts, I calculate a 3ma unaccounted for current draw from something.
So I have same conclusion, one or more of the following is wrong.
510 Ohm too high, 5 volts too low, un-shown current draw thru Leakey caps or extra circuits or bad data.
(plus the pot is (and should be) more likely a 10 K ohm)
ws
******************
----- Original Message -----
From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
Sent: Friday, July 03, 2009 11:58 PM
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
> Warren
>
> The Vcc dc current doesn't change much if the circuit is oscillating (at
> least it shouldn't if the circuit operating correctly - the active
> devices aren't supposed to saturate as thats adds too much phase noise).
>
> My simulations were over the 50-100C rang although temperature doesn't
> affect the currents and voltages too much.
> They also don't include loading of the frequency control pot wiper.
>
> There is no major inconsistency other than the effect of the
> unanticipated loading of the EFC pot wiper.
> The simulated oscillator Vcc is actually about 3V when everything is
> working properly.
> NB its important to use models that accurately reflect the correct
> transistor junction area just blindly tossing in 2N3904's increases the
> current somewhat as the Vbe is lower for a given current than the actual
> transistors. The simulated oscillator Vcc is then closer to 2.5V.
>
> Earlier circuit schematics didn't include as many actual measured dc
> voltages.
>
> Bruce
>
> WarrenS wrote:
>> Bruce
>>
>> 4ma? Interesting, Maybe IF the Osc was running, BUT
>> My DC calculations of the present values I can read on Schematics come to 1.3ma + 0.26 = 1.58 ma load on the 510 ohm
>>
>> With it NOT osc, which is almost the present case,
>> the Only Dc current drawn is.
>>
>> 1.75 V on 20K = 0.08ma (Q1 bias)
>> 1.04 V on 2.2K = 0.47ma (Q1 collector)
>> 0.365V on 470 ohm. = 0.77 ma (Q2 emitter)
>> Osc board =1.32 ma
>>
>> Buffer board
>> (2*4.7k) at 2.49 V = 0.26 ma
>>
>>
>> Also note that the 4.7 K ohm pot and 4.7K divider voltages show excess current draw somewhere, maybe on the Grn lead bypass cap..
>>
>> ws
>>
>> **************
>>
>> ----- Original Message -----
>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>> Sent: Friday, July 03, 2009 11:08 PM
>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>
>>
>>
>>> Warren
>>>
>>> My Spice model of the dc levels in the oscillator indicate a current
>>> more like 4mA in the 510 ohm resistor.
>>>
>>> Bruce
>>>
>>> WarrenS wrote:
>>>
>>>> I just saw your schematic of the rest of the buffer circuit. Sorry I missed it early.
>>>>
>>>> The Osc Vcc is shown driven from a 510 Ohm resistor, and with the total current thru the 510 res at under 2 ma,
>>>> Looks like Osc Vcc should be at more like 4.0 volts instead of 2.49 volts.
>>>> If I'm correct about that (this time) then, It has got to be one of three things.
>>>> The 510 ohm is wrong, the 5 volts is wrong, or the load on the 510 ohm is wrong.
>>>> It should be easy enough to measure which one it is.
>>>> The voltage at the junction of the two 4.7K resistors are WRONG.
>>>> Looks like it is caused by an excess current draw on the Grn lead.
>>>> Check the direction and leakage of the voltage controlled cap diode and bypass.
>>>> If all else is right, Look for the cause of the extra loading current on the 510 ohm,
>>>>
>>>> WS
>>>>
>>>> ****************
>>>>
>>>> ----- Original Message -----
>>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>>> Sent: Friday, July 03, 2009 9:55 PM
>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>
>>>>
>>>>
>>>>
>>>>> WarrenS wrote:
>>>>>
>>>>>
>>>>>> A couple more thoughts;
>>>>>>
>>>>>> Is the VCC (Red wire) current draw correct?.
>>>>>> Should be 1.3 ma (per the schematic values) with the 470 Ohm and
>>>>>> more like 7 plus ma with the 47 ohm.
>>>>>>
>>>>>> AND I don't remember seeing, IS the Freq real close AND can it be tuned by the cap and Green wire voltage?
>>>>>> If both answers are YES, at least most things are working correctly.
>>>>>>
>>>>>> As Bruce said:
>>>>>>
>>>>>>
>>>>>>
>>>>>>> "if the resistor values are correct this is the voltage(s) that one should get"
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>> With the resistors values shown, I agree that all the other voltages are correct given the 2.49 volts,
>>>>>> BUT the 2.49 volt is NOT controlled by any of the values on that schematic, it's just an input.
>>>>>> AND the output amplitude is not yet being controlled by current cut off.
>>>>>>
>>>>>> Something that needs to be asked is are all these the correct values.
>>>>>> It sounds like you got them by measuring and not from a "should be" schematic.
>>>>>>
>>>>>> One last, way out thought.
>>>>>> It was my belief that very good osc have some sort of AGC to control the amplitude,
>>>>>> ANY chance that the VCC is used as the input to control the amplitude of the Osc output?
>>>>>> If so, could it be that the external AGC that is what broken and supplying too low a VCC voltage to the circuit?
>>>>>>
>>>>>> ws
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>> The Colpitts crystal oscillator limits the crystal current by being
>>>>> periodically cutoff for part of the cycle.
>>>>> There is an optimum duty cycle (usually something like 25%) that
>>>>> corresponds to minimum phase noise.
>>>>> The oscillator dc collector current determines the crystal current and
>>>>> hence the output amplitude.
>>>>> It is possible to perform AGC by varying the oscillator Vcc but the
>>>>> output amplifier schematic appears to have no circuitry for this.
>>>>> Some circuits actually do this, but the oscillator transistor is still
>>>>> cutoff for part of the oscillator cycle.
>>>>> The oscillator VCC is fed from a tap in a resistive divider network
>>>>> (located on the external (to the oven) buffer board) connected between
>>>>> the 5V supply and ground.
>>>>> Usually one just varies the emitter (or collector) dc current by
>>>>> selecting a resistor value.
>>>>>
>>>>> Bruce
>>>>>
>>>>>
>>>>>> *****************
>>>>>> ----- Original Message -----
>>>>>> From: "Bruce Griffiths" <bruce.griffiths@xtra.co.nz>
>>>>>> To: "Discussion of precise time and frequency measurement" <time-nuts@febo.com>
>>>>>> Sent: Friday, July 03, 2009 8:21 PM
>>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>
>>>>>>> WarrenS wrote:
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> Here is my two cents worth
>>>>>>>> 20 mv output, sure sounds like something is broken.
>>>>>>>> It should be fixed before it is modified.
>>>>>>>> The 2.49 volts on the Red input voltage seem LOW, More Vcc maybe.
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>> Not so, if the resistor values are correct this is the voltage that one
>>>>>>> should get.
>>>>>>> The question that remains: Is the voltage what the designer intended?
>>>>>>> With the Colpitts oscillator used limiting in the oscillator transistor
>>>>>>> (usually by current cutoff) is used to limit the amplitude.
>>>>>>> Just increasing the voltage without checking the resultant crystal
>>>>>>> current will be within permissable limits can be counter productive.
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> The "Grn" labeled wire, FreqCtrl input should be about 1/2 VCC for testing.
>>>>>>>> If you do 'need' to modify the gain,
>>>>>>>> It would seem better to bypass the 470 ohm resistor with a cap in series with the 47 ohms.
>>>>>>>>
>>>>>>>> ws
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>> Bruce
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> ********************
>>>>>>>> ----- Original Message -----
>>>>>>>> From: "Mike Monett" <xde-l2g3@myamail.com>
>>>>>>>> To: <time-nuts@febo.com>
>>>>>>>> Sent: Friday, July 03, 2009 11:50 AM
>>>>>>>> Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>
>>>>>>>>> > An update on the Broken Ovenaire OSC 85-50.
>>>>>>>>>
>>>>>>>>> > I prepared a 'schematic' of the Output Board and the Oscillator
>>>>>>>>> > Board (attached) and I have lots of pictures of the external unit
>>>>>>>>> > and the insides if anyone is interested.
>>>>>>>>>
>>>>>>>>> > I resoldered all connections and replaced all transistors on the
>>>>>>>>> > Output Board and the Oscillator Board all to no benefit. I
>>>>>>>>> > measured all the components with an LCR meter and found the 0.01
>>>>>>>>> > uF bypass on the 330 ohm resistor in the emitter circuit of the
>>>>>>>>> > output transistor of the Output Board to be low and with a high
>>>>>>>>> > ESR. I replaced this with about a 20% increase in output amplitude
>>>>>>>>> > but still inadequate. I replaced the rest of the 0.01 uF caps on
>>>>>>>>> > the output board with no additional benefit. I transiently
>>>>>>>>> > disconnected the Red wires from the Oven Controller board and
>>>>>>>>> > there was no increase in output or significant increase in voltage
>>>>>>>>> > to the Oscillator Board.
>>>>>>>>>
>>>>>>>>> > Therefore, it appeared that a 'low output crystal' (if such a
>>>>>>>>> > thing exists) was the only logical explanation that I could come
>>>>>>>>> > up with. That seeming to be the case, there appeared to be only 4
>>>>>>>>> > options. 1. Toss the OCXO (sorry, too much effort so far). 2.
>>>>>>>>>
>>>>>>>>> > Build an external amplifier (seemingly too much additional
>>>>>>>>> > effort). 3. Try to adjust on the bias of the oscillator transistor
>>>>>>>>> > to achieve a higher output (seemed too 'iffy'). Or 4. Lower the
>>>>>>>>> > value of the resistor in the emitter circuit of the Oscillator
>>>>>>>>> > Board to get more gain out of the last stage in the Oscillator
>>>>>>>>> > Board.
>>>>>>>>>
>>>>>>>>> > I replaced the 470 ohm resistor with a 47 ohm resistor and the
>>>>>>>>> > amplitude increased to about 0.4 V P-P into a 50 ohm load and was
>>>>>>>>> > sufficient to make it a usable OCXO again.
>>>>>>>>>
>>>>>>>>> > I reassembled, resealed with Epoxy and all seems well so far.
>>>>>>>>>
>>>>>>>>> > If anyone wants pictures or other info, please let me know.
>>>>>>>>>
>>>>>>>>> > Thanks for all the suggestions and help.
>>>>>>>>>
>>>>>>>>> > Joe
>>>>>>>>>
>>>>>>>>> Joe,
>>>>>>>>>
>>>>>>>>> Congratulations on getting your system to work!
>>>>>>>>>
>>>>>>>>> A couple of things. First, trying to measure the currents in the
>>>>>>>>> circuit with a ferrite toroid won't do you much good. You don't know
>>>>>>>>> what the currents should be, and the secondary of the toroid
>>>>>>>>> transformer requires a termination resistor. The value changes with
>>>>>>>>> the turns ratio.
>>>>>>>>>
>>>>>>>>> Just from looking at the circuit, the RF currents will be extremely
>>>>>>>>> low. This requires a large number of turns on the secondary, which
>>>>>>>>> will probably resonate at or below the 10MHz operating frequency due
>>>>>>>>> to stray capacitance from the connection to the scope. So it is
>>>>>>>>> unlikely you will get any useful progress in this direction.
>>>>>>>>>
>>>>>>>>> However, from the values on your schematic, the output tank circuit
>>>>>>>>> resonates at 9.602MHz with a Q of 9.6. So the tank is already well
>>>>>>>>> below resonance, which attenuates the output voltage.
>>>>>>>>>
>>>>>>>>> Any stray capacitance you add to the circuit will bring the resonant
>>>>>>>>> frequency lower, further aggravating the loss in signal.
>>>>>>>>>
>>>>>>>>> The output tank is tapped with the 75pF and 91pF in series. This
>>>>>>>>> further attenuates the signal.
>>>>>>>>>
>>>>>>>>> I'd change the circuit to a single capacitor across the tank with a
>>>>>>>>> small trim capacitor to tune it to resonance.
>>>>>>>>>
>>>>>>>>> To get the signal into 50 ohms for distribution, I'd add a limiter
>>>>>>>>> if you can tolerate a square wave output, or a good emitter follower
>>>>>>>>> if you need a sine wave. Take the output from the collector of the
>>>>>>>>> 2N2369 to get the maximum signal amplitude.
>>>>>>>>>
>>>>>>>>> Your original post mentions an output amplitude of 20mV. If the
>>>>>>>>> normal amplitude is around 2V, this represents a loss of 40dB. This
>>>>>>>>> is a huge loss in signal. The circuit obviously worked at one time,
>>>>>>>>> so there may well be some other hidden problem.
>>>>>>>>>
>>>>>>>>> It is possible the crystal is damaged, but this seems unlikely. A
>>>>>>>>> crystal oscillator probably won't even start if the signal level is
>>>>>>>>> down 40dB.
>>>>>>>>>
>>>>>>>>> You can check the oscillator and crystal in SPICE. Normally, the
>>>>>>>>> high Q of the crystal will make the analysis very slow. It could
>>>>>>>>> take many hours for the simulation to begin oscillating and
>>>>>>>>> stabilize at the final amplitude. The transient analysis requires a
>>>>>>>>> very fine time step for accuracy, and you could run out of memory
>>>>>>>>> before the simulation was complete.
>>>>>>>>>
>>>>>>>>> I have developed a much faster way of analyzing a crystal oscillator
>>>>>>>>> in SPICE. Instead of requiring tens or hundreds of thousands of
>>>>>>>>> simulated cycles, this method gives accurate results in only a few
>>>>>>>>> dozen cycles. For more information, please see "SPICE Analysis of
>>>>>>>>> Crystal Oscillators"
>>>>>>>>>
>>>>>>>>> http://pstca.com/spice/xtal/clapp.htm
>>>>>>>>>
>>>>>>>>> You can estimate the value of the crystal ESR by finding the Q of
>>>>>>>>> your crystal and working backwards.
>>>>>>>>>
>>>>>>>>> I'm attaching a gif of your schematic for reference. This is rotated
>>>>>>>>> 90 degrees and enhanced in LView Pro to improve the contrast.
>>>>>>>>>
>>>>>>>>> Please let me know if you have any questions.
>>>>>>>>>
>>>>>>>>> Thanks,
>>>>>>>>>
>>>>>>>>> Mike
>>>>>>>>>